Entry  Mon Feb 3 17:15:59 2020, Duo, DailyProgress, Electronics, Some confusing noisemon findings 12x
    Reply  Tue Feb 4 17:30:50 2020, Duo, DailyProgress, Electronics, Some confusing noisemon findings p.pdfDataAndCode.zipPhoto_on_2-4-20_at_5.27_PM.jpgnoisemon_v1.pdf
Message ID: 1845     Entry time: Mon Feb 3 17:15:59 2020     Reply to this: 1846
Author: Duo 
Type: DailyProgress 
Category: Electronics 
Subject: Some confusing noisemon findings 

Trying to figure out the nature of the distortion found in noisemon (1843), I made a piece of DSUB9 breakout like attachment 1. I just used two wires and two clippers wires to break the DSUB connector into something I can clip into the board.

I sent a 10Hz drive to the coil driver and measured the spectrum at all the test points (including one pair on the coil driver board).

First, I compared the distortion situation at the input and output of noisemon (attachment 2). I measured the spectrum at TP10 and TP12 of coil driver (this is where noisemon picks up the differential input, reference DCC: D070483), and subtracted them quadratically, which gives us the input signal of noisemon. Then I measured the spectrum at the output of noisemon. Looking at attachment 2, it seems the coil driver does not contribute to the distortion.

Then I used the probe and measured all the test points individually. Attachment 3 also shows TP10 and TP12 but instead of measuring them individually and subtracting them, I just put one probe on TP10 and another on TP12. It looks like the coil driver does have a little distortion, contrast to the conclusion above.

Attachment 4 shows the spectrum at the noisemon input. There is a buffer between TP10/TP12 on coil driver and the input of noisemon (reference D070483). It looks like the distortion is less, somehow.

Attachment 5 is the spectrum after the passive filters, between TP3 and TP4. It seems the distortion comes back a little under 60Hz.

Attachment 6 shows the spectrum after the instrument amplifier.

Attachment 7 and 8 shows the output of two stages of high pass filters. The distortions gets much worse after two HP filters.

Attachment 9 shows the output of the low pass filter. We can see the high frequency harmonics are gone.

Attachment 10 is very confusing. Between TP8 and TP9 is just a buffer (LT1128 buffer unfortunately) - why does it give so many lines even without any drive? I tried to see them in the oscilloscope but there wasn't anything significant. Maybe it was the clippers/wires making the noise? I did check the signal at the conjunction point between the clipper and the wire using oscilloscope (attachment 11) - it is indeed bad (attachment 12) - but why these lines only show up in TP9?

I am very confused and need to think about what is going on now. I think a couple immediate questions are:

1. Why do I see so many lines at the output of noisemon even when there is no input? I should only see noisemon noise when there is no drive, but I did not short the inputs like when I measure the noise. I did shorted the inputs and checked the noise after the measurement - it is normal. Thus, the lines could be caused by not shorting the inputs of the coil driver, but I did not short the inputs in other measurements either - I just unchecked the excitation on diaggui.

2. Besides the TP9 confusion, we still see a lot of harmonics in uppersteam testpoints. Considering the TP9 situation, I think we should first ask - are they real? Then, if they are real, how bad are they? Considerations could include the functionality of noisemon board at the sites? This potentially include risk of saturation, increase of the noisemon noise and errors in measuring the DAC noise.

Attachment 1: thumbnail_image0.jpg  266 kB  | Hide | Hide all
thumbnail_image0.jpg
Attachment 2: distortionDiagnosis.pdf  32 kB  | Hide | Hide all
distortionDiagnosis.pdf
Attachment 3: CoilDriverOutput.pdf  15 kB  | Hide | Hide all
CoilDriverOutput.pdf
Attachment 4: NoisemonInput.pdf  15 kB  | Hide | Hide all
NoisemonInput.pdf
Attachment 5: NoisemonPassiveFilter.pdf  15 kB  | Hide | Hide all
NoisemonPassiveFilter.pdf
Attachment 6: NoisemonTP5.pdf  15 kB  | Hide | Hide all
NoisemonTP5.pdf
Attachment 7: NoisemonTP6.pdf  15 kB  | Hide | Hide all
NoisemonTP6.pdf
Attachment 8: NoisemonTP7.pdf  15 kB  | Hide | Hide all
NoisemonTP7.pdf
Attachment 9: NoisemonTP8.pdf  15 kB  | Hide | Hide all
NoisemonTP8.pdf
Attachment 10: NoisemonTP9.pdf  16 kB  | Hide | Hide all
NoisemonTP9.pdf
Attachment 11: IMG_1685.jpg  608 kB  | Hide | Hide all
IMG_1685.jpg
Attachment 12: IMG_1684.jpg  720 kB  Uploaded Mon Feb 3 18:16:17 2020  | Hide | Hide all
IMG_1684.jpg