  637   Thu Mar 4 23:03:14 2010 DmassLaserDoublingMZ Complete!

I have finished rebuilding the Mach Zehnder. There is some serious scatter/glow from various places, but there was before as well. I am dumping all the obvious beams on blades. Pictures incoming.

I am coloring 3-5" in height on the inside of the box with sharpie, in an attempt to absorb the scatter around beam height. I have no idea if this change in color actually corresponds to changing the reflectivity for 1064, but I made an attempt at a blackbody anyways.

Full post + pix to come

  636   Thu Mar 4 22:33:15 2010 ranaLaserGYROFinesse measurement

I'm sure that its more precise to just scan the counter propagating beam with the first direction locked (or not). In that case, you get to directly read the frequency off of the thing driving the AOM. No cal required.

 

  635   Thu Mar 4 22:18:05 2010 ZachLaserGYROFinesse measurement

Whoa, that seems high! In the post I made about this new setup, I estimated the finesse to be ~550 assuming 50 ppm losses and no degradation in R for the Y1S mirrors in our cleanest of clean labs. Of course, the value we are using for the PZT gain may be off by a factor of 2 or more, so we'll know better once we do what you're saying with the sidebands. For completeness, what was the drive signal to the PZT for this measurement, 10 Vpp triangle @ 100 Hz?

Quote:

I scanned the cavity to measure the FWHM of the transmitted carrier.  The plot is attached.  I have converted from volts to frequency using 2.5MHz/V as the piezo response.  This number might be entirely made up, and I will actually try to calibrate this tomorrow using the method that Frank explained (using the sidebands as a frequency reference).

The FWHM from this measurement was 79704Hz.  I measured the cavity length and got 3.175m, but this is probably accurate to around +-4cm.  It is really hard to measure right up to the mirror, particularly when they're in different heights of mirror mount.  Having said that, Frank thinks the error in the voltage measured by the scope could be up around 10% so that will probably be the dominant error in our finesse measurement.

This gives a finesse of 1185 and a FSR of 94.4MHz.

 

  634   Thu Mar 4 20:41:42 2010 AlastairLaserGYROFinesse measurement

I scanned the cavity to measure the FWHM of the transmitted carrier.  The plot is attached.  I have converted from volts to frequency using 2.5MHz/V as the piezo response.  This number might be entirely made up, and I will actually try to calibrate this tomorrow using the method that Frank explained (using the sidebands as a frequency reference).

The FWHM from this measurement was 79704Hz.  I measured the cavity length and got 3.175m, but this is probably accurate to around +-4cm.  It is really hard to measure right up to the mirror, particularly when they're in different heights of mirror mount.  Having said that, Frank thinks the error in the voltage measured by the scope could be up around 10% so that will probably be the dominant error in our finesse measurement.

This gives a finesse of 1185 and a FSR of 94.4MHz.

Attachment 1: GYRO_bandwidth.jpg
GYRO_bandwidth.jpg
  633   Thu Mar 4 01:38:40 2010 ranaElectronicsGYROUPDH box s/n 1437 examined

After some hunting, I found the correct schematic for this Universal PDH box. It is v1 of the D0901351 variety. The 'version numbers' that Alberto refers to in the Wiki are his own made up numbers and don't correspond to the actual numbers on the schematics.

Attached is the LISO model results of what we can get with ~10 R and C changes. I found that luckily the zero in the U7 stage, as stuffed, happens to cancel the cavity pole frequency of the gyro! Seems too lucky to be true, but that's what the numbers say. I await for Alastair to measure the cavity finesse and bandwidth for real in the morning.

In this new PDH circuit design, I've made the boost stage have a high frequency gain of -1 instead of -0.01. The boost frequency is getting moved up to ~1-2 kHz. It will be a true integrator below this frequency.

The OP27 input stage will have a gain of 40 instead of 2. This is good for noise. By using a 25 Ohm resistor on the input, this will also balance the current in the 2 legs of the OP27 and reduce some offset.

We will also take away the 50 Ohm resistor at the OP27 (U1) input. The proper place to terminate the IF port of the mixer is at the IF port and NOT after the IF low pass filter. Those mini-circuits mixers are not 50 Ohms above the bandpass frequencies. That makes the high harmonics shoot backwards into the mixer and make it angry.

According to my LCR meter, the NPRO PZT capacitance is 8.5 nF (as measured from the front panel of the controller). So its to small to be used to make the dominant pole in the servo. We'll keep the pole in the front of the circuit.

Soldering to follow in the morning...

Attachment 1: u.pdf
u.pdf u.pdf
  632   Thu Mar 4 01:26:50 2010 ranaElectronicsDoublingGreenPD Noise

Unimportant EMI - ignore and press on at full speed.

  631   Wed Mar 3 17:58:29 2010 DmassElectronicsDoublingGreenPD Noise

I built the attached circuit twice on a 3M solderless protoboard, and had a noise discreprency between my two channels when trying to measure the PD dark noise.

I plugged the PD into a scope, and saw the following difference (second picture). These ringdowns occur whether or not my PD is on.

What is this op amp behavior, and what makes it worse?

The scale is the same for both traces on the scope.

Attachment 1: GreenPD.png
GreenPD.png
Attachment 2: photo(3).jpg
photo(3).jpg
  630   Wed Mar 3 15:49:09 2010 FrankComputingDAQhard disk replaced

i've replaced the temporary hard disk for the full frames by a new, larger one. As nobody wants data from the busted one i will send it to the manufacturer for an exchange now...

  629   Wed Mar 3 08:39:41 2010 ranaElectronicsGeneralOP AMP WOES

Woe.  The noise of the 50 Ohm resistor is ~1 nV/rHz. The gain of your opamp is 100k/50 = 2000. So the output noise should be at least 2 uV/rHz.

Its better to diagnose the noise by not having the resistor there and/or to put the PD there. I have also added the LT1792 to the LISO SVN.

  628   Wed Mar 3 00:22:05 2010 DmassElectronicsGeneralOP AMP WOES

I tried to make a simple current to voltage convert for my Green PDs, and measure the noise. Hilarity did not ensue.

I made the following connections for an LT1792 (I also tried a LT1012).

 

            |-----100pF----|          

            | ___ 100k_____|__   

            |    |\           |   

Gnd-50Ohm-BNC----|-  \ _______|___BNC____ to SR785

Gnd--------------|+  /

                 |/               

I stuffed this into my existing hydra board, and pulled out all the other components. My Noise looked WAY TOO HIGH

I was getting well over a uV/rtHz up to 800 Hz. This does not add up.

  • Johnson Noise should be 40 nV/rtHz
  • Voltage noise of the 1792 is plotted to be 4nV/rtHz with a 30 Hz corner
  • Current Noise of the 1792 is said to be 10 fA/rtHz (which gives 1nV/rtHz)
  • This does not add up to microvolts!

I also tried:

  • Changing the Capacitor values to 22 pF and 220 pF, neither seemed to fix anything.
  • Pulling both the R and the C from the feedback, replacing it with a 3k resistor, and still about 1uV....
  • Changed in the LT1792 for a LT1012 (which had similar values for I and V noise, so this doesn't add up either).
  • Checked the terminals of the op amps for power, I was sending +15V to 7 and -15V to 4.
  • Yes I double checked that the 785 was not in a stupid gain state, and that the noise floor of the measurement was way below this.

I am curious about:

  • My voltage supply: I stole an unused new focus +/- 15V supply from the shelf above the GYRO, I don't know if this could be causing problems to that extent.
  • I tried two versions of a 3M board, with the same disappointing results.
  • I also tried a TENMA 72-6615 "Laboratory DC Power supply" It is regulated to 0.01% according to its data sheet. Maybe it is the connection scheme I am using?
  • The transfer function, I will take this tomorrow and check to see that it is A/B=1

There really has to be something stupid I am doing...

  627   Tue Mar 2 21:03:36 2010 KojiLaserGYROcavity "locked" in both directions

Congratulation! This is an epoch! 

Quote:

 I was able to get the cavity to stay locked in the unmodulated (CCW) direction by increasing the gain of the slow channel. The issues we were having this morning were caused by the ~25 kHz resonance of the PZT, which must have reached unity gain when we put enough PDH gain in. The added slow gain seems to be enough at low frequency to keep the thing locked, but if I try to up the fast gain the thing starts to oscillate. We will have to modify the servo to accommodate the PZT spike to make the lock more robust.

Once having locked the original direction, I steered the double-passed beam from the AOM into the CW direction of the cavity and aligned it to isolate a 00 mode at the output. It's not quite perfect, as the REFL signal is still clipped a bit out of the faraday isolator, but it's enough to couple a decent amount of power into the cavity, and the clipping will likely disappear when we put (any) modematching optics in this path.

The AOM seems to be working very well. I was able to ramp the frequency tuning voltage into the driver and see the different modes scan through on the CCD at the output. I increased the voltage from 7.8 V (which corresponded to 45 MHz) to 8.77 V (which must be very close to the desired FSR/2 of 47.5 MHz), and saw a strong 00 at the output when the CCW beam was on resonance. Hence, the CW beam is not strictly on resonance, but it is "locked" a fixed frequency very close to one FSR from the CCW beam, and we see something close to resonance at the output as a result.

The input power in each direction is close to the same (as cursorily seen on an IR card), and the emergent power at the transmitted end of the CW path is roughly half of that on the CCW path. Given NO modematching in the CW direction, this must mean that we are fairly close to resonance in that direction.

Tomorrow, we will begin doing the following:

  • Work out some at least crumby modematching for the CW beam
  • Adapt the second bluebox into a tentatively working form for the other loop
  • Set up the REFL_CW readout and requisite RF electronics
  • Lock the CW beam
  • Pilfer a Marconi
  • Beat the signals at the output
  • Have a working gyro
  • Celebrate with champagne and whisky

Here is a picture of both directions simultaneously outputting a reasonable 00 with a macroscopic frequency shift ~1 FSR:

 

  626   Tue Mar 2 20:09:16 2010 ZachLaserGYROcavity "locked" in both directions

 I was able to get the cavity to stay locked in the unmodulated (CCW) direction by increasing the gain of the slow channel. The issues we were having this morning were caused by the ~25 kHz resonance of the PZT, which must have reached unity gain when we put enough PDH gain in. The added slow gain seems to be enough at low frequency to keep the thing locked, but if I try to up the fast gain the thing starts to oscillate. We will have to modify the servo to accommodate the PZT spike to make the lock more robust.

Once having locked the original direction, I steered the double-passed beam from the AOM into the CW direction of the cavity and aligned it to isolate a 00 mode at the output. It's not quite perfect, as the REFL signal is still clipped a bit out of the faraday isolator, but it's enough to couple a decent amount of power into the cavity, and the clipping will likely disappear when we put (any) modematching optics in this path.

The AOM seems to be working very well. I was able to ramp the frequency tuning voltage into the driver and see the different modes scan through on the CCD at the output. I increased the voltage from 7.8 V (which corresponded to 45 MHz) to 8.77 V (which must be very close to the desired FSR/2 of 47.5 MHz), and saw a strong 00 at the output when the CCW beam was on resonance. Hence, the CW beam is not strictly on resonance, but it is "locked" a fixed frequency very close to one FSR from the CCW beam, and we see something close to resonance at the output as a result.

The input power in each direction is close to the same (as cursorily seen on an IR card), and the emergent power at the transmitted end of the CW path is roughly half of that on the CCW path. Given NO modematching in the CW direction, this must mean that we are fairly close to resonance in that direction.

Tomorrow, we will begin doing the following:

  • Work out some at least crumby modematching for the CW beam
  • Adapt the second bluebox into a tentatively working form for the other loop
  • Set up the REFL_CW readout and requisite RF electronics
  • Lock the CW beam
  • Pilfer a Marconi
  • Beat the signals at the output
  • Have a working gyro
  • Celebrate with champagne and whisky

Here is a picture of both directions simultaneously outputting a reasonable 00 with a macroscopic frequency shift ~1 FSR:

bidirectional_lock_ish.jpg 

  625   Tue Mar 2 12:41:23 2010 DmassElectronicsDoublingOp Amps

Here are plots for the noise of my op amps (in Volts).

I compared 8 op amps, these are the two winners I came up with, AD743 clearly wins. I also included (for later reference) the noise of the ISS PDs I am using.

Op amps I looked at:

  • AD: 711 743 797 829
  • OP: 27 
  • OPA: 627
  • LTI: 1128 1028

This list was made out of Frank's head based on what he thinks we already have. The AD743 seems to give a SNR of 1k throughout my measurement band, which was my somewhat arbitrary requirement (e.g. based on what I got optically when I used prisms, but I don't know how much my old layout was limiting me).

 

Attachment 1: GreenAD743z100k.PNG
GreenAD743z100k.PNG
Attachment 2: GreenOPA627z100k.PNG
GreenOPA627z100k.PNG
Attachment 3: ISSAD829z500.PNG
ISSAD829z500.PNG
  624   Mon Mar 1 19:41:25 2010 ZachLaserGYROAOM double-pass

 We were able to resolve the AOM double-pass issue by enlarging the beam going into it. We had originally tried using the smallest practical waist (~ 70 um) because the jitter suppression scheme only works perfectly--i.e. beam ROC = mirror ROC with the mirror exactly R away from the AOM--in the limit where zR --> 0, or w0 = 0. It appears that this caused the beam to diverge too quickly within the AOM, which is pretty long. We have changed the waist size to ~200 um at the expense of moving the mirror in to z = 15.5 cm from the AOM (as opposed to z = R = 30 cm as it was before). We are now able to isolate a decent-looking double-pass beam, which we verified by measuring its power. Hopefully the loss in jitter suppression will be insignificant as we only expect the AOM to shift by hundreds of Hz or so, corresponding to a very small angular displacement.

After this progress this morning, I tried to realign the cavity and get the original unmodulated direction to lock, but I am having quite some difficulty. I can align it and isolate the 00 mode pretty well, and there appears to be a decent amount of power emerging at the transmitted end on resonance (as seen on IR card), but the thing just doesn't want to lock. I have tried to adjust the RF phase to no avail. I suspect it has something to do with a polarization issue with the BS I am using to screen some power from the REFL beam before the PD, but I can't figure it out. I will try again tomorrow morning.

Once this direction locks again, we can feed in the modulated beam around the other direction, and we should have all the necessary equipment to set up the second loop. I spoke with Frank today, and I believe we will need a Marconi to set up a PLL at the output to do the frequency readout. We will figure this out when it comes up, I guess.

  623   Fri Feb 26 16:19:05 2010 DmassLab InfrastructureDoublingDoubling Experiment Update!

Of course, one might ask why I rebuilt the Mach Zehnder readout...

Here are plots of the various PDs with one arm blocked and the HEPA on (I lated checked that the HEPA being off didn't really fix the coherence). Things to note:

  • The black trace is the frequency domain subtraction of the two signals from one wavelength. The subtraction agrees with the subtraction predicted by the coherence (not shown).
  • The PD noise is very much not limiting my measurement, however, the coherence was terrible.
  • I think this had something to do with the huge path length between my prism and my PDs, which could make jitter couple in very strongly.
  • The 532 noise limits were "bad", but is consistent with what you expect from the current noise of an AD829 coupling through a 50k transimpedance resistor (using LISO).

In conclusion, I think I was geometry limited before (using prisms), and I am rebuilding the readout with an SNR of 1000. This is more than 3, so its better

Op Amps:

  • I want to use AD743s for the I to V converter in my Si PDs, they have decent voltage noise with a 100k transimpedance resistor.
  • (I also looked at the AD829 and confirmed that it is fine with a 500 Ohm transimpedance resistor at DC).
  • Plots to come as I play with gnuplot
Attachment 1: PD1AnaNoise.pdf
PD1AnaNoise.pdf
Attachment 2: PD2AnaNoise.pdf
PD2AnaNoise.pdf
Attachment 3: PD3AnaNoise.pdf
PD3AnaNoise.pdf
Attachment 4: PD4AnaNoise.pdf
PD4AnaNoise.pdf
  622   Fri Feb 26 01:50:06 2010 DmassLab InfrastructureDoublingDoubling Experiment Update!

Some more updates:

  • The "random mirror" is said to be a super mirror by Frank. It is some sort of splitter, so not interesting to me. I labeled it (mounted) and put it on the top right optics shelf.
  • Frank helped me out with another REO supermirror. It had some dust on it, so I blew it off. I did not wipe it; it looked clean.
  • I measured the 532 transmission to be 56 and 58% for the two mirrors, so I will use these to separate my light.
  • I adjusted my mode matching so that I had a 400 um (width) spot at my green PDs (no extra lenses!).
  • I am using some lenses to get the right spot size for my ISS PDs.
  • Pictures/Diagrams soon!
  621   Thu Feb 25 01:00:31 2010 DmassLab InfrastructureDoublingDoubling Experiment Update!

Update on rebuilding:

MIRRORS

  • Rana found me a 4 mirror REO package with 2 mirrors in it.
  • One of these is a supermirror.
  • The other is something unknown that looks different, has an engraved barrel, and is most definitely NOT high reflective (maybe a splitter) at 1064.
  • I can't find the 1064 transmission (of 100 mW) of the supermirror on a viewer, or sensitive card. It appears to be very small. This is good.
  • I see bright 532 spots reflected from both surfaces of this mirror. This is less than good.
  • I still have 100 uW transmitted, which is within a factor a 3 of my max power. This will give me around 1V DC with a 50K load resistor on one of the PDs I will use (below).
  • I am using some 90% R 45P (the beam is S) to dump a large fraction of my IR power before my PDs.

PDs:

  • I took one of the ISS PDs off the 40m PSL table. Siobhan was her name.
  • I had another ISS PD from Stefan's time here.
  • I pulled the 500 ohm SM transimpedance resistor from both of these, and replaced it with a 500 Ohm thin film resistor from the 40m. (I didn't get a good pic, but had no problems soldering, and Koji gave me a tired thumbs up when I asked him if I "did it wrong")
  • I am going to use the ISS PD box to power and readout these.
  • I also have two DET110's  and two nice and old Thorlabs PDs of unknown model number (but known PD type: Hamamatsu S1223).
  • The DET110's have a quoted NEP of 2e-14 W/rtHz and the Hamamatsu's have a quoted NEP of 1e-14.
  • These numbers are way better than the newer thorlabs PDs I was using (which were total shite - like 10 pA/rtHz NEPs of 1e-11 to 1e-12. Terrible.)
  • I will think about the loading tomorrow, and see what I can get away with for each of these. The batteries are 12 and 22.5V, respectively.
  • A 150K load would fill the DAQ with what I have, so maybe this is a good number to shoot for?

Table Layout:

  • I think I have everything fitting!

Beam Dumps:

  • Frank found some of the Si slabs we are using for the low scatter beam dumps, I am planning on fashioning two crude dumps from these (I have only 3 slabs).
  • I have up to 200 mW coming out of my MZ, so I have to dump some so as not to cook the ISS PDs.
  620   Wed Feb 24 04:35:04 2010 DmassLaserDoublingNew Mach Zehnder Layout!

I shrank the Mach Zehnder by 2" per side to accommodate putting the PD readout inside THE BOX. It is still tight, but I think I can come up with a way to fit everything.

 

The Mach Zehnder is somewhat realigned, and somewhat mode matched (the fringes arent obviously two different modes anymore). Once I put PDs in, I can fine tune the alignment easily.

  619   Wed Feb 24 00:27:14 2010 DmassLab InfrastructureDoublingDoubling Crystal Damaged

I am in the process of rebuilding the Mach Zehnder with the following goal:

I want the entire experiment to live inside one box, readout included. To this end I needed to compactify the Mach Zehnder somewhat.

In realigning the ovens, I looked closer at the crystal which was giving me problems before (the exact same behavior). When I look through it, I can see what looks like damage through the center of the crystal. This is consistent with the crazy behavior I get where aligning through the center of the crystal produces terrible beam profiles.

SHG CRYSTAL:

______________________________________________

|                   //                       |

|                 //  <--- Crack?            |

|_______________//___________________________|

 

I will proceed with the experiment, having to go with some funny alignment.

  618   Mon Feb 22 15:44:47 2010 FrankComputingDAQday(s) for data rescue?

which days do we need to copy? It's a lot more data as expected, 64gb per day (!) so plz let me know which day is required to copy. We should time it well as the framebuilder will delete the old data almost instantly to free space for tyhe new data. So if we delete all current data it depends on how much of the old stuff we copy how long we have to access the data before its deleted automatically

  617   Mon Feb 22 13:24:04 2010 AlastairLaserGYROupdate (AOM)

Quote:

Quote:

 Alastair and I have been working on getting the AOM double-pass setup going. We have borrowed a curved Y1 from Peter to use as a retroreflector while we wait for the one we ordered. We are using a Tektronix FG through the 2-W MiniCircuits RF amp on the shelf to drive the AOM at the moment, and that appears to work decently well. We are running into issues with alignment, however.

Right now, we can get the 1st-order beam isolated with an iris and reflected back along the same path back into the AOM, but what comes back out isn't very clean. What we should see is the singly diffracted beam that gets reflected straight back through the AOM (without a second diffraction), and then the doubly diffracted beam--what we want--which should be a little dimmer and should lie exactly along the input beam. We tried lowering the AOM drive power until only one beam was left (the single-pass beam), but this never happened; all we got was something that looked like two dim beams close together or one very elliptical one.

We think it is because we don't have any way of adjusting the orientation of the AOM at the moment. Alastair has drawn up a new mount which will allow us to put a tilt stage on top. In the meantime, we will try to get something remotely reasonable out and attempt to lock the cavity with that.

Yes, the mount for the AOM sucks ass.  I drew up a new one last night that will incorporate one of the New Focus 9082 alignment stages.  It's with the machine shop now so we should have it towards the end of next week.  I also ordered the alignment stage from Newportfocusnew (as I will now be calling them).

 That's the 9082 stage arrived now.

  616   Mon Feb 22 13:14:26 2010 FrankComputingDAQfb0 replacement disks arrived

the new disk for the framebuilder arrived. So i try to move the data from the last days before the crash to the disk which is currently installed. As i was asked to copy only the last view days the space on this smaller disk is sufficient. If everything is done i will partition the new, larger disk and copy everything to there. Is there any time i can't shut down the frontend? Any plans for working with the RT stuff so far? If not, i'll try to do this late today or tomorrow afternoon, depending on the work on the cavity stuff...

  615   Fri Feb 19 10:22:26 2010 AlastairLaserGYROupdate (AOM)

Quote:

 Alastair and I have been working on getting the AOM double-pass setup going. We have borrowed a curved Y1 from Peter to use as a retroreflector while we wait for the one we ordered. We are using a Tektronix FG through the 2-W MiniCircuits RF amp on the shelf to drive the AOM at the moment, and that appears to work decently well. We are running into issues with alignment, however.

Right now, we can get the 1st-order beam isolated with an iris and reflected back along the same path back into the AOM, but what comes back out isn't very clean. What we should see is the singly diffracted beam that gets reflected straight back through the AOM (without a second diffraction), and then the doubly diffracted beam--what we want--which should be a little dimmer and should lie exactly along the input beam. We tried lowering the AOM drive power until only one beam was left (the single-pass beam), but this never happened; all we got was something that looked like two dim beams close together or one very elliptical one.

We think it is because we don't have any way of adjusting the orientation of the AOM at the moment. Alastair has drawn up a new mount which will allow us to put a tilt stage on top. In the meantime, we will try to get something remotely reasonable out and attempt to lock the cavity with that.

Yes, the mount for the AOM sucks ass.  I drew up a new one last night that will incorporate one of the New Focus 9082 alignment stages.  It's with the machine shop now so we should have it towards the end of next week.  I also ordered the alignment stage from Newportfocusnew (as I will now be calling them).

Attachment 1: AOM_mount.pdf
AOM_mount.pdf AOM_mount.pdf
  614   Fri Feb 19 07:54:14 2010 ranaElectronicsDoublingBad PD?!

Quote:

I will be using a Thorlabs DET110 with something like a 5 or 10k load resistor, and an SR560 for gain to replace the really crappy thorlabs PD for green.

 Now I really wish we had made those low noise DCPDs...I guess that you can steal the ISS PDs for now. Just move them into the appropriate place and use the ISS box to power them and read them out. We can just continue to use some Thorlabs diode + SR560 as an ISS for now.

  613   Fri Feb 19 01:04:35 2010 DmassElectronicsDoublingBad PD?!

I will be using a Thorlabs DET110 with something like a 5 or 10k load resistor, and an SR560 for gain to replace the really crappy thorlabs PD for green.

  612   Thu Feb 18 21:16:01 2010 DmassElectronicsDoublingBad PD?!

Quote:

One of my switchable gain Si detectors appears to have particularly bad noise perfomance.

I compared the dark Voltage noise of my Thorlabs PDs. I am barely below the ADC noise on one of my PDs, and it onlyu gets worse at lower frequencies, so for Beta, on my current setup, the PD noise is limiting me below about 5 Hz.

 

Detector Noise (V/rtHz) @ 10 Hz Gain Setting
Alpha(532) 600 n 47500 V/A
Beta(532) 4 u 47500 V/A
Gamma(1064) 40 n 4750 V/A
Delta(1064) 200  n 4750 V/A

 

 Added a measurement:

Attachment 1: PDDarkNoise.pdf
PDDarkNoise.pdf
  611   Thu Feb 18 20:09:32 2010 MottLaserFiberMode Shape of the fiber laser

Quote:

 I made a rough measurement of the mode shape of the laser used in the fiber setup.  I need to go back in to get more points and estimate the error, but to first approximation the beam waist is about 1550 microns.  The data I was able to take is included below (the x axis is in cm, the y in microns), but I will go back and take more data tomorrow morning.

 

It's also nice to put the curve that you're fitting with, the fitting parameters, and to extend the plotted curve to show the waist itself

  610   Thu Feb 18 18:49:21 2010 DmassElectronicsDoublingBad PD?!

One of my switchable gain Si detectors appears to have particularly bad noise perfomance.

I compared the dark Voltage noise of my Thorlabs PDs. I am barely below the ADC noise on one of my PDs, and it onlyu gets worse at lower frequencies, so for Beta, on my current setup, the PD noise is limiting me below about 5 Hz.

 

Detector Noise (V/rtHz) @ 10 Hz Gain Setting
Alpha(532) 600 n 47500 V/A
Beta(532) 4 u 47500 V/A
Gamma(1064) 40 n 4750 V/A
Delta(1064) 200  n 4750 V/A

 

  609   Thu Feb 18 18:14:20 2010 ZachLaserGYROupdate (AOM)

 Alastair and I have been working on getting the AOM double-pass setup going. We have borrowed a curved Y1 from Peter to use as a retroreflector while we wait for the one we ordered. We are using a Tektronix FG through the 2-W MiniCircuits RF amp on the shelf to drive the AOM at the moment, and that appears to work decently well. We are running into issues with alignment, however.

Right now, we can get the 1st-order beam isolated with an iris and reflected back along the same path back into the AOM, but what comes back out isn't very clean. What we should see is the singly diffracted beam that gets reflected straight back through the AOM (without a second diffraction), and then the doubly diffracted beam--what we want--which should be a little dimmer and should lie exactly along the input beam. We tried lowering the AOM drive power until only one beam was left (the single-pass beam), but this never happened; all we got was something that looked like two dim beams close together or one very elliptical one.

We think it is because we don't have any way of adjusting the orientation of the AOM at the moment. Alastair has drawn up a new mount which will allow us to put a tilt stage on top. In the meantime, we will try to get something remotely reasonable out and attempt to lock the cavity with that.

  608   Thu Feb 18 18:04:07 2010 MottLaserFiberMode Shape of the fiber laser

 I made a rough measurement of the mode shape of the laser used in the fiber setup.  I need to go back in to get more points and estimate the error, but to first approximation the beam waist is about 1550 microns.  The data I was able to take is included below (the x axis is in cm, the y in microns), but I will go back and take more data tomorrow morning.

Attachment 1: BeamMode_Laser_Rough.png
BeamMode_Laser_Rough.png
  607   Thu Feb 18 04:29:02 2010 DmassLaserDoublingAA Noise, and slow channels added...

Some Plots:

  1. The noise in the various channels (in counts) with 50 ohm terminators at the input to the AA box.
  2. The Spectra (in V) for each PD with alternate arms blocked, and measured AA noise plotted (lower curve).
  3. The coherence between same color PDs for the two different arms being blocked - Things look as expected based on the power levels.

More to come.

 I made analog measurements of the PD dark noise, and for the most part it seems below the noise downstream of the AA box.

BAFFLING BEHAVIOR SEE PLOT TITLED MZWTF:

When I block one arm, I have decent coherence above 1 Hz. When I block the other, the same is true. When I unblock both arms, and let the Mach Zehnder interfere, my Coherence takes a dive at 1-2 Hz. I DO NOT HAVE ANY IDEA HOW THIS IS POSSIBLE.

THE INDIVIDUAL TIME SERIES LOOK LIKE I EXPECT, AND THE MACH ZEHNDER IS STILL ALIGNED ENOUGH TO INTERFERE.

Attachment 1: AntiAliasNoiseALL.pdf
AntiAliasNoiseALL.pdf
Attachment 2: ArmBlockSpectra.pdf
ArmBlockSpectra.pdf
Attachment 3: ArmBlockedCoh.pdf
ArmBlockedCoh.pdf
Attachment 4: WTFMZ.pdf
WTFMZ.pdf
  606   Thu Feb 18 00:45:53 2010 DmassLaserDoublingAA Noise, and slow channels added...

I started chasing my PD noise, and I found extra noise after the input to the AA box, on two channels, one of which I was using in my Mach Zehnder setup.

  • Channels 26 and 30 show extra noise by more than a factor of 3 below 1 Hz.
  • I will be using channels 23, 24, 28, and 29 for recording my data now.

I also added some slow channels to the DAQ.

  • I downsampled my data to 256 via the DAQ (I believe it uses these filters)
  •  
  • The elliptic filters used for downsampling have some high frequency flatenning
  • If I don't do a digital lowpassing of my signal, all my high frequency noise will be aliased into my measurement band many (256) times, which causes it to increase by a factor of 10, despite being filtered.
  • I added an 8th order Butterworth low pass filter at 256 Hz (sample rate, not Nyquist) via Foton to kill these aliased contributions without touching the downsampling filter shape too much.
  • I added the gain conversion to Volts, so my 256 Hz channels are now in V!

 These are name as follows in the DAQ:

Table Name Channel 4 kHz Name (Defunct for now) 256 Hz Name
Alpha (Green PD 1) 23 C2:ATF-MZ_DC1_IN C2:ATF-MZ_ALPHA256
Beta (Green PD 2) 24 C2:ATF-MZ_DC2_IN C2:ATF-MZ_BETA256
Gamme (IR PD 3) 28 C2:ATF-MZ_AC2_IN C2:ATF-MZ_GAMMA256
Delta (IR PD 4) 29 C2:ATF-MZ_AC3_IN C2:ATF-MZ_DELTA256

 

  605   Wed Feb 17 17:46:56 2010 DmassLaserDoublingIntensity Noise Measurement, and Coherence

It seems that one of my photodiodes was bad. PD4 (the IR phodotiode that was incoherent with all the others) was swapped for a PD of the same model from Thorlabs, and the mystery incoherence disappeared.

I still am only at the 95-99% coherence level at the DAQ below 8 Hz, but before PD4 was only 70% Coherent with all the other PDs, and it was the only PD that had this behavior.

I added a plot of what I am working with rright now. DC1 and DC2 are green, DC3 and DC4 are IR. The Spectra are taken with 1 arm blocked. The pink curve is my ADC noise.

 

Attachment 1: PDCoherence.pdf
PDCoherence.pdf
  604   Wed Feb 17 03:40:10 2010 DmassComputingDoublingMATLAB MZ SUBTRACTION CODE

I have updated the 40m SVN with all the code needed to run the MZ subtraction. The data files are about 250 megs total. I will upload them to the elog if that is acceptable, if not I will upload the downsampled data (which should be about 16 megs).

I put the transfer function inside the subtraction function (because I was doing it wrong before). Now f_domainsubtraction.m should be useful as a general frequency domain subtraction tool.

Attachment 1: MZ_Code_Diagram.png
MZ_Code_Diagram.png
  603   Wed Feb 17 01:27:24 2010 FrankComputingGeneralfb0 is now working again

Quote:

as it turned out that one of the hard disks failed so i had to replace it. The device contains all "full frame" data, nothing else. Unfortunately i didn't have a spare disk of that size so i replaced it by a smaller one. So the new volume is empty and no past (full) data is available. Trend data is OK.

So if nobody needs the full data from the last couple of weeks i will send the disk back to WD to get it replaced. If the data is/was important we can wait until i get a new disk of the same size (i ordered one today). If this shows up the next two days or so we could try to copy most of the full data to the new one, but only if really required as it takes ~6h or so to duplicate the disk and i would like to avoid setting everything up. So if anyone needs the old full data plz let me know. If i don't here something within the next two days i will send the broken disk back to WD. Again: any TREND data is good, only FULL data is broken.

 No data I have taken needs to be recovered. If it costs us *very little effort/money, I would like to have the last few days, but I can retake it all too.

  602   Tue Feb 16 19:28:25 2010 FrankComputingGeneralfb0 is now working again

as it turned out that one of the hard disks failed so i had to replace it. The device contains all "full frame" data, nothing else. Unfortunately i didn't have a spare disk of that size so i replaced it by a smaller one. So the new volume is empty and no past (full) data is available. Trend data is OK.

So if nobody needs the full data from the last couple of weeks i will send the disk back to WD to get it replaced. If the data is/was important we can wait until i get a new disk of the same size (i ordered one today). If this shows up the next two days or so we could try to copy most of the full data to the new one, but only if really required as it takes ~6h or so to duplicate the disk and i would like to avoid setting everything up. So if anyone needs the old full data plz let me know. If i don't here something within the next two days i will send the broken disk back to WD. Again: any TREND data is good, only FULL data is broken.

  601   Tue Feb 16 16:54:58 2010 AlastairLaserGYROCurved mirror for AOM

I just ordered a replacement curved mirror for double passing the AOM.  The one we have at the moment is for 45P, and we really need 0 degrees incidence.  It is listed as in stock and I've put down for quick delivery.  It will be a Y1-1025-0-0.30CC

  600   Tue Feb 16 13:21:08 2010 AlastairLaserGYROBeam scan for AOM double pass

 I've taken a beamscan for the path where the AOM is going to be installed.  This is just so that we can accurately pick the lens and curved mirror that are going in here.  There will be a second mode matching telescope after this set of optics before the CW beam goes into the cavity.

Attached are:

1) Matlab figure of the fit

2) the text file of the beam scan (1st column is z distance from last turning mirror (m), 2nd column is x-diameter (um), 3rd column is y-diameter (um))

3) Photo showing the location on the bench of the 'last turning mirror' prior to the AOM that the z values are taken relative to (in other words the mirror is position z=0)

 

From the fit the values for the waist radius and location are:

x  267+- 4um,  at  -0.324 +- 0.009 m

y 356 +- 3 um, at -0.277 +- 0.004m

 We will fit the optics for the AOM using the average of these which is a waist of 311.5um at 0.3005m

Attachment 1: 10_02_16_beamscan_for_AOM.jpg
10_02_16_beamscan_for_AOM.jpg
Attachment 2: 10_02_16_beam_pre_AOM.txt
0.0365	1086.9	940.5
0.0465	1106.3	964.7
0.0565	1123.1	981
0.0665	1145.5	982.4
0.0765	1163.1	983.7
0.0865	1189.6	990.5
0.0965	1208.5	985.5
0.1065	1231.7	1020.2
0.1165	1237.3	1026
0.1265	1266.2	1047.7
0.1365	1286.2	1050.4
0.1465	1316.3	1069.1
0.1565	1334.7	1079.6
0.1665	1353.4	1110.8
0.1765	1384.1	1116.8
0.1865	1388.2	1126.8
0.1965	1425.9	1139.6
0.2065	1448.8	1152.5
0.2165	1468.6	1165.5
0.2265	1488.1	1182.4
0.2365	1492.2	1196
0.2465	1531	1213.6
0.2565	1549.2	1244.2
0.2665	1574.1	1264.1
0.2765	1598.3	1281.4
0.2865	1646.2	1305.3
0.2965	1647.7	1308.3
0.3065	1687.9	1313.5
0.3165	1718.3	1340.8
0.3265	1716.3	1342.9
0.3365	1739	1370.1
0.3465	1783.7	1385.8
0.3565	1793.2	1399.6
0.3665	1798.4	1410.8
0.3765	1822.6	1423.6
0.3865	1860.1	1441.3
0.3965	1882.4	1463.6
0.4065	1901.1	1479.6
0.4165	1916.5	1493.2
0.4265	1941.4	1510.1
0.4365	2013.7	1543.2
0.4465	2040.4	1551.2
0.4565	2079.9	1570.1
0.4665	2090.1	1597.4
0.4765	2126.7	1605.5
0.4865	2157	1629.6
0.4965	2183	1634.9
0.5065	2201.1	1643.7
0.5165	2230.6	1674.6
0.5265	2276.1	1687
0.5365	2271	1703.2
Attachment 3: Screen_shot_2010-02-16_at_1.34.37_PM.png
Screen_shot_2010-02-16_at_1.34.37_PM.png
  599   Tue Feb 16 11:39:38 2010 FrankComputingGeneralfb0 error

we had a disk failure on the weekend on one of the harddisks. So fb0 will be down until we replaced that disk...

  598   Tue Feb 16 03:32:58 2010 dmassComputingDoublingMatlab MZ Subtraction Code Update

I have added all my subtraction code to the SVN, and will try to keep it under version control. It should also now be comprehensible to anyone who wants to read it. It is in this directory on the 40m svn:

/svn/trunk/mDV/extra/C2/dmass/mzdata/

I have included a .png diagram of the code.

  597   Sun Feb 14 04:31:09 2010 DmassComputingDoublingMATLAB MZ SUBTRACTION CODE

Quote:

Do you detrend x1-x4 before you calculate phi532 and phi1064? I thought detrending changes the DC values and thus leads wrong results for the phases.

Quote:

I have diagrammed my MZ subtraction algorithm. I am redoing my code now that I have realized a few things so that another person can read it. I will update the guide with the associated MATLAB code names, and continue to put everything in the SVN.

 

x1,x2,x3,x4 are the time series of my PD's during a segment of free running MZ data.

n1,...,n4 is a (detrended) time series of my noise in each PD due to some source.

 

 oops, I do not. Changing diagram to reflect this.

  596   Sat Feb 13 13:29:22 2010 KojiComputingDoublingMATLAB MZ SUBTRACTION CODE

  594   Thu Feb 11 19:07:09 2010 DmassLaserDoublingIntensity Noise Measurement, and Coherence

Koji and I tried to chase down some incoherence in my intenstiy noise measurement.

Setup:

  • Block one arm, look at signals on all four PDs

Problem:

  • Very low coherence between my IR PDs

Things wrong that we found:

  • The IR beam was misaligned on both IR PDs. Maybe that alignment of the prisms drifted? I think their mounting is not so good (I use a CVI BS mount, but there are no set screws to lock it). The path after this is also long, so small changes to the prism angle could easily toss me off the PD.
  • The signal itself not coherent between the two IR PDs with one arm blocked!! We tracked it down to one PD having a huge excess noise, which was scatter.  I  dump most of the power onto a razor blade dump before the PD (For MZ_DC4_IN - the second IR PD), and by changing the angle of the razor blade dump, and turning off the HEP /(and lights), we were able to change the coherence drastically.

THE PLOT (SUMMARY)

  • DC1 and DC2 are my Green PDs. DC3 and DC4 are my IR PDs. I assume there are still scatter issues. For the red/blue the HEPA was on. The improvement we got by just turning the HEPA off was less than the pink/green curves. The blade dumps themselves seems really crappy. The surface is not black to look at, and when you move the beam around on them, the intensity (as seen on a viewer) changes a lot.
Attachment 1: MZPDCoherence.pdf
MZPDCoherence.pdf
  593   Thu Feb 11 16:40:27 2010 AlastairLaserGYROMore stuff for gyro

 We now have the mount for the AOM.  Zach and I went looking for a suitable curved mirror to double pass the AOM and found a 30cm 45P at the 40m.  I'm guessing it's not ideal since we're using normal incidence but we'll probably have to order another one.

Zach had thought out the jitter side of things, and in order to ideally get rid of this we want the distance from the AOM to match the radius of curvature of the mirror.  Then we need to set the waist size so the wavefront curvature matches the mirror.

In order to do this we need Zr<<Radius of mirror.  In other words we want a small waist in the AOM.  Somewhere around 70um should allow us to have the waist on the way back to be about 1mm different in position from on the way through if we're using the 30cm mirror.

  592   Thu Feb 11 15:19:44 2010 MottLaserFiberFiber Mach-Zehnder

 The inner loop of the mach-zehnder is setup and crudely aligned. The next step is to finish the alignment, lock the small mach-zehnder and then build the bigger one around it.  

  591   Wed Feb 10 08:29:38 2010 ranaLaserDoublingNoise Spectra (Done)

Quote:

 

Also: I am still baffled by why I can't subtract better where there is such a high coherence at low frequency, it's above 0.99, yet I only seem to be able to get a 1.2ish order subtraction. My frequency domain results are about the same as the Wiener filtering results generated by the code: MZwino in the svn.

 You should use your knowledge of mathematics to make an estimate for the amount of coherent signal based upon the coherence estimate. Then you can plot the predicted subtraction factor v. the actual subtraction.

  590   Wed Feb 10 08:27:08 2010 ranaLaserGYROFisheye

We need a fisheye scheme so's that we can stick a camera underneath the shelf and take a picture of the whole gyro side of the table. Ideas?

  589   Tue Feb 9 23:27:45 2010 DmassLab InfrastructureOpticsWindow Fun

There were some windows (W2 CVI parts) laying around the ATF. I measured the zero degree transmission for 1064 and 532 by sticking them in the output path of the Mach Zehnder (between the Mach Zehnder box and the PD Readout box).

I left the cover on the PD Readout box, so am not positive that I did not misalign the beams onto the PDs, though I was using at roughly normal incidence (less than 10 degrees), so I doubt it.

Results are for normal incidence, which is not what the coatings were designed for in every case. Beam polarization here should be P.

 

Part T_1064 T_532
W2-PW-1025-C-1064-0 91% 89%
W2-IF-1025-UV-1064-45P 8% (not a typo)* 91%
W2-IF-1025-C-1064-45S 81% 97%

I played around with the second one a bit to see if I could get the IR to go through. It did not seem to work. Maybe a user error due to lateness? Maybe a previous user error in labeling? Maybe CVIs coatings just happened to do that! Looks like I won't use a 45P window at 0 deg!

  588   Tue Feb 9 17:58:15 2010 ZachLaserGYROGyro cavity locked in other direction

 The gyro cavity is now aligned and lockable in either direction (we are still waiting for a mount for the AOM, so we haven't done both at the same time yet!). We have the input beam split and going through both faradays, and the REFL beam in each goes through the other's on the way out. The faradays have been configured so that the REFL beams are rejected on the way back through, and we are able to lock either direction by hooking up the PDH to the corresponding one. We also have CCDs and TRANS PDs for both directions at the far end.

The first picture below just shows the cavity locked in the new (clockwise) direction. The second is a diagram of the input/REFL system (NOTE: In the diagram, there is no PD at REFL CCW. This is because we only have one supply-powerable PDA255 at the moment. The others are battery powered, but Alastair is converting another one for permanent use).

The next thing we're going to do is move the input beamsplitter back a ways to make room for the AOM double-pass setup. We are going to put this together with the mounts we have now just to work out the spacing, etc.

lock_shot_small.jpg

input_diagram_small.jpg

  587   Tue Feb 9 17:55:17 2010 AlbertoMiscGeneralMarconi missing

Quote:

Quote:

the Marconi from the ATF is missing. Does someone know where it is? We need it NOW back, not tonight, not tomorrow, NOW !!!

I have it.

I want a million in dime coins and a helicopter ready to leave from the 40m.

---

I legally borrowed it last week from the ATF. I'm bringing it back to you.

 sorry :-) i didn't know about your deal. i asked Mott and was looking for it too so i thought someone forgot to bring it back after he borrowed it. i need it only for the next couple of hours...