Message ID: 811     Entry time: Fri Jun 11 20:33:17 2010     Reply to this: 815
Author: Dmass 
Type: Computing 
Category: DAQ 
Subject: Attempt at oven transfer functions 

The plot I attached and didn't explain is:

  • Green = Voltage across the Ohm Ranger as driven and read out by the (unknown) bridge of the Newport 3040
  • Red = Voltage across a high power 5 Ohm resistor I am using on the current output of the Newport 3040
  • Units are standard DAQ conversion ~ 1600 counts / Volt

Notes:

  • I use an SR560 to buffer the signal from the ohm ranger before inputting it into the AA chassis
  • The resistance between the inner BNC leads of the AA box was ~2 kOhms - FOR SEVERAL DIFFERENT CHANNELS
  • I have the PI (no D) servo set at its lowest settings, so lowest integral gain, and lowest proportional gain.
  • I make step responses in the green curve via the ohm ranger.
  • I should see step responses of the red from the proportional term, and changes in the slope of the integration following these step responses.
  • I see SLOPE IN THE GREEN. This is weird.
  • The slope in the green goes away when I unplug the voltage (from the TEC drive current) across the 5 ohm resistor from the front end.
  • The slope in the green stops when the current rails.
  • I don't understand the coupling, but since I don't have circuit diagrams for the 3040 - I stopped thinking about it, and just used a USB stick in a scope to take this data.

Moral:

  • The 3040/ front end acts funny when you try to read out the current and the voltage of the thermometer.
  • I don't really care that this is the case.
  • The front end seems to be working close to all the way
Attachment 1: dmassFE.png  14 kB  | Hide | Hide all
dmassFE.png