40m
QIL
Cryo_Lab
CTN
SUS_Lab
TCS_Lab
OMC_Lab
CRIME_Lab
FEA
ENG_Labs
OptContFac
Mariner
WBEEShop
|
Cryo Lab eLog |
Not logged in |
 |
|
Fri Apr 30 12:07:25 2021, aaron, DailyProgress, Laser, marconi locking
|
Fri Apr 30 15:15:33 2021, rana, DailyProgress, Laser, marconi locking
|
Mon May 3 23:28:56 2021, rana, DailyProgress, Laser, delay line frequency discriminator
|
Wed May 5 11:54:19 2021, aaron, DailyProgress, Laser, delay line frequency discriminator 9x
|
Wed May 5 22:21:17 2021, rana, DailyProgress, General, delay line frequency discriminator
|
Thu May 6 10:48:40 2021, aaron, DailyProgress, Laser, delay line frequency discriminator 
|
Tue May 11 14:37:48 2021, aaron, DailyProgress, Laser, delay line frequency discriminator, current noise measurement
|
Wed May 12 15:35:09 2021, aaron, DailyProgress, Laser, delay line in a box  
|
Wed May 12 16:52:08 2021, aaron, DailyProgress, Laser, delay line frequency discriminator   
|
Thu May 13 16:31:40 2021, aaron, DailyProgress, Laser, delay line frequency discriminator 6x
|
Thu May 13 19:09:16 2021, rana, DailyProgress, Laser, delay line frequency discriminator
|
Fri May 14 18:43:25 2021, aaron, DailyProgress, Laser, delay line frequency discriminator
|
Thu May 20 12:50:56 2021, aaron, DailyProgress, Laser, delay line frequency discriminator
|
Fri May 21 14:15:21 2021, aaron, DailyProgress, Laser, current noise of custom current drivers  
|
Tue Aug 17 14:04:27 2021, aaron, DailyProgress, Laser, delay line frequency discriminator
|
Wed Aug 18 09:55:13 2021, aaron, DailyProgress, Laser, delay line frequency discriminator  
|
Thu Aug 19 14:21:35 2021, aaron, DailyProgress, Laser, delay line frequency discriminator  
|
Fri Aug 20 11:03:44 2021, aaron, DailyProgress, Laser, delay line frequency discriminator   
|
Mon Aug 23 13:38:57 2021, aaron, DailyProgress, Laser, delay line frequency discriminator
|
Tue Aug 24 16:44:39 2021, aaron, DailyProgress, Laser, delay line frequency discriminator  
|
Thu Aug 26 11:58:50 2021, aaron, DailyProgress, Laser, delay line frequency discriminator 9x
|
Fri Aug 27 15:27:38 2021, rana, DailyProgress, Laser, delay line frequency discriminator
|
Tue Aug 31 15:14:28 2021, aaron, DailyProgress, Laser, quieting the Rio W and Teraxion beat notes, but not much help   
|
Wed Sep 1 12:20:18 2021, rana, DailyProgress, Laser, quieting the Rio W and Teraxion beat notes, but not much help
|
Thu Aug 19 16:32:47 2021, rana, DailyProgress, Laser, delay line frequency discriminator
|
Tue May 4 10:50:22 2021, aaron, DailyProgress, Laser, marconi locking 
|
Wed May 5 00:27:10 2021, rana, DailyProgress, Laser, marconi locking
|
|
Message ID: 2797
Entry time: Fri Aug 20 11:03:44 2021
In reply to: 2795
Reply to this: 2798
|
Author: |
aaron |
Type: |
DailyProgress |
Category: |
Laser |
Subject: |
delay line frequency discriminator |
|
|
I'm driving the input of the delay line box with 10 dBm from Marconi's RF output.
- First, tuned the Marconi carrier frequency to 112.8 MHz to null the output of the delay line mixer (after DC coupled, G=1, dynamic range mode, SR560 buffer with 1 kHz, 6 dBm/oct lowpass).
- Next, frequency modulating the RF carrier at 314.14 Hz with 800 kHz deviation from the Marconi. The output of the delay line box is an 16 mVpp sine wave, see attachment 1
- I checked out the noise spectrum of the Marconi buffered through both the DC coupled SR560 above, and separately an AC-coupled, G=100, low noise mode SR560 with 10 kHz, 6 dBm/oct lowpass
- On the Moku, the spectra look reasonable. The G=1 and G=100 peaks at 314 Hz differ by 40 dB, as expected. Attachment 2.
- Feeding the signal to our ADC has some issues. Not sure what's going wrong (ground loops? hot swapped AA-ADC cable?), but any signal I plug in to the ADC results in the periodic noise in attachment 3 on not only the signal-carrying channel but also the adjacent channels. The discontinuities are larger than the G=1 buffered signal, though not as large at the G=100 buffered signal.
- I'm using the PSOMA channels, so the names are innaccurate. In attachment 3, the DC-coupled signal is in SLD_PDH_CTL, while the AC-coupled signal is in SLD_PDH_SIG
- The noise is at about 120 Hz, which suggests some power line issue. Measuring only the AC-coupled SR560 in battery powered mode, with no other instruments connected after the buffer, improves but does not eliminate the noise (attachment 4). Also, on battery powered mode, the SR560 constantly overloads; my exact sequence of steps to observe this overload behavior was (1) power off the SR560, (2) unplug the line voltage, (3) power on the SR560.
|
|
|
|
|
|
|
|
|