i'm currently adding new channels for RFAM monitoring and EOM temp control but before they work i used some other channels for now.
- The EOM temp setpoint is currently modulated using the SR785 as i cant add a large DC offset on a small modulation using the SR345.
- The current setpoint is 650mV which is equivalent to something around 30C.
- On top i have a modulation (sine, 1mHz) of 180mV (?, have to double check, might be 170mV) to modulate the EOM temperature.
- The setpoint value is recorded using channel C3:PSL-FSS_VCOMON
- The actual Temperature of the in-loop sensor is recorded using C3:PSL-FSS_FREQCOUNT
- The feedback to the laser is monitored using the fast actuator channel C3:PSL-FSS_FAST. The EOM feedback loop is disabled as well as the script for laser temp feedback.
i checked the temp modulation by using a PT1000 sensor as an OOL sensor on the EOM case using a multimeter.with min/max hold function.
- min value is 1120 Ohms = 30.9C
- max value is 1139 Ohms = 35.8C
Monitoring the the above channels using Striptool didn't show any result yet. Even the modulation of ~5K is not visible in the feedback to the laser. Will add all signals to their final channels tomorrow and do a more detailed analysis. |