Mon Mar 25 22:56:42 2013, Evan, DailyProgress, Electronics Equipment, Frequency readout noise of PDH loop   
|
Tue Mar 26 10:00:19 2013, Evan, DailyProgress, Electronics Equipment, Frequency readout noise of PDH loop
|
Thu Mar 28 12:36:27 2013, Zach, DailyProgress, Electronics Equipment, Frequency readout noise of PDH loop
|
Sat Mar 30 01:42:16 2013, tara, DailyProgress, Electronics Equipment, Frequency readout noise of PDH loop
|
|
Message ID: 1132
Entry time: Thu Mar 28 12:36:27 2013
In reply to: 1128
Reply to this: 1133
|
Author: |
Zach |
Type: |
DailyProgress |
Category: |
Electronics Equipment |
Subject: |
Frequency readout noise of PDH loop |
|
|
Is this really the input noise of the PDH servo (i.e., output noise / servo TF)? If so, it seems pretty high. With the right components, you should to be able to do ~100x better.
Quote: |
Tara was able to tame the cavity servo loop. The third attachment is the result of several SR785 measurements of the error signal power spectrum. I converted this to a frequency noise power spectrum (fourth attachment) by extracting the voltage/frequency calibration factor from the error signal as follows.
|
|