  • With both electrodes driven at a positive voltage, the results are still an improvement over the original design, but by smaller factors, particularly in 3 of the higher frequency modes at around 2 and 3 kHz. With another parametric sweep I may be able to find a better design. The opposite voltage was useful because it could couple force in opposite directions to adjacent anti-nodes with opposing signs. An adjusted configuration could probably be found to interact with anti nodes with the same signs. An alternative option would be to leave the radial position of the more central ESD constant but rotate it relative to the sample by some amount, though this would also require an additional parametric sweep as well as a much larger ESD. 