Author: Gabriele 
Type: Optics 
Category: Daily Progress 
Subject: Quality factor measurement of a Mark Optics disk at LMA 

[Massimo Granata (LMA), Quentin Cassar (LMA), Gabriele] 

This week I'm visiting LMA to learn how their Gentle Nodal Suspension system works and to measure the quality factors (Q) of one of Mark Optics disks. First of all we annealed the disk for 9 hours at 900 degrees (plus 9 hours warm up and 9 hours cool down).

Then we installed the disk into the measurement system and started by searching for all the resonances.

My COMSOL simulation proved to be good enough to give us the frequencies, especiallty after a small fine tuning of the disk thickness (within specs). We identifies a total of 32 modes of different families, and measured the ring down of all of them. Since our disk has no flats, each mode is actually a doublet with very small frequency separation. The analysis software has a bandwith of 1 Hz to find the peak amplitude, so it can't resolve the two modes. When both are excited to a significant amplitude by the electrostatic actuator, we see a clear beat in the ring-down. I had to write a new fitting code to take this into account. More details will follow in a DCC document. However, here I can say that the fit works remarkably well for all modes. 

A couple of examples:

 

Here is a summary plot of the quality factor and loss angle for all modes. We measured Q as high as 10e6, in line with other LMA samples (2") we tested in these days. In conclusion, the Mark Optics disks, as they are, are good enough for our coating tests.

   