  5295   Wed Aug 24 11:30:27 2011 jamie, jenne, kiwamu, suresh, steveUpdateSUSETMX wiped, replaced, door on

We've closed up ETMX:

  • the optic was drag wiped
  • the suspension tower was put back in place
  • earthquake stops were backed off the appropriate number of turns, and de-ionized
  • chamber door was put on
  11730   Wed Nov 4 15:48:58 2015 SteveUpdatePEMETMX wirestadoff vs EQ

ETMX was not much effected by this 8.3 M Chilian earthquake. It was kicked up and it was kicking for hours after it. This is a sign of instability but it maintaind it's position.

The 24 hours plot follows the tempeerature.

 

  5061   Fri Jul 29 16:04:51 2011 steveUpdateSUSETMX, PRM & SRM OSEM gains set

Jamei fixed the computer. Now I had a healthier ETMX with SIDE signal that allowed me to set gains to get Q of 5

The Y arm restore locked the arm at TRY -out 2.4V,    the X arm still did not lock or align to lock. It has some fringes.

ETMX  POS gain   22  -> 17,              PIT  3  ->  1,                    YAW  3,                     SIDE     -170    ->    -110

PRM                        50  ->  30,                     1,                                       2,                                       50

SRM                        55  ->  20,                    2.4  ->  1.5,                      4.8  ->  2,                          140   ->    100

Now all SUS OSEM gains are set

  17147   Tue Sep 20 18:18:07 2022 AnchalSummarySUSETMX, ETMY damping loop gain tuning

[Paco, Anchal]

We performed local damping loop tuning for ETMs today to ensure all damping loops' OLTF has a Q of 3-5. To do so, we applied a step on C1:SUS-ETMX/Y_POS/PIT/YAW_OFFSET, and looked for oscillations in the error point of damping loops (C1:SUS-EMTX/Y_SUSPOS/PIT/YAW_IN1) and increased or decreased gains until we saw 3-5 oscillations before the error signal stabilizes to a new value. For side loop, we applied offset at C1:SUS-ETMX/Y_SDCOIL_OFFSET and looked at C1:SUS-ETMX/Y_SUSSIDE_IN1. Following are the changes in damping gains:

C1:SUS-ETMX_SUSPOS_GAIN : 61.939   ->   61.939
C1:SUS-ETMX_SUSPIT_GAIN :   4.129     ->   4.129
C1:SUS-ETMX_SUSYAW_GAIN : 2.065     ->   7.0
C1:SUS-ETMX_SUSSIDE_GAIN : 37.953  ->   50

C1:SUS-ETMY_SUSPOS_GAIN : 150.0     ->   41.0
C1:SUS-ETMY_SUSPIT_GAIN :   30.0       ->   6.0
C1:SUS-ETMY_SUSYAW_GAIN : 30.0       ->   6.0
C1:SUS-ETMY_SUSSIDE_GAIN : 50.0      ->   300.0

 

  3925   Mon Nov 15 16:09:44 2010 JenneUpdateVACETMX, ETMY, ITMY chambers opened

[Steve, Jenne, Suresh, Koji]

The remaining test mass chambers have been opened, and have light doors in place.  Now we can do all of the rest of the IFO alignment, and then (hopefully) button up before the New Year.

  7326   Fri Aug 31 10:16:02 2012 janosch, SteveUpdate ETMX, scattering preps

The alignment of the pick-off mirror near ETMX is done. Everything turned out to be easy once we realized that there is no sense getting the alignment laser (going through viewport to pick-off to ITMX) back to ETMX. It is only necessary to hit ITMX somehow, since this makes sure that there is one scattered beam that will make it from ITMX to pick-off through viewport.

After the auxiliary optic (that we never used in the end) was removed again, we levelled the optical table.

So in the current setup, we can have small-angle scattering measurements on ITMX and large-angle scattering measurements on ETMX.

  12068   Fri Apr 8 09:13:20 2016 SteveUpdateendtable upgradeETMX-T beam height

X arm resonating after alignment, beam height on ETMX optical table ~4.75"

 

  4792   Mon Jun 6 23:56:16 2011 rana, valeraConfigurationSUSETMX/ETMY OSEM whitening

 We measured the OSEM PD whitening transfer function of the ETMX OSEM UL whitening stage (D000210) by comparing the input signal to the whitening amplifier (single pin LEMO monitor) to the output signal - both were piped into the DAQ. The transfer function was close to constant 0 dB/180 deg independent of the whitening switch selection (FM1 filter engaged/disengaged)  up to ~20 Hz where we run out of coherence. All other ETMX and ETMY spectra at the input of the digital whitening compensation don't change when the whitening is switched on/off so by induction we conclude that all the ETMX/ETMY OSEM PD hardware whitening filters are not on.

  4833   Fri Jun 17 17:02:15 2011 JamieUpdateSUSETMX/ETMY binary output modules (re)installed, not yet tested

I have installed a new binary output module in ETMY, where there was none previously.  It is installed, powered (with working LEDs), hooked up (to the binary output card and the cross connect), but it hasn't been fully tested yet.

I also re-installed the binary output module in ETMX, with newly modified power-indicator LEDs.

Both modules are fully installed, but they have not yet been fully tested to confirm that they are indeed switching the whitening and de-whitening filters.

  4683   Tue May 10 23:14:01 2011 kiwamuUpdateSUSETMX_OPLEV : He-Ne laser dead ?

I found that a He-Ne laser which has been used for ETMX_OPLEV was NOT giving the light.

Since I didn't find the switch key for it I have no idea if the laser is simply off or dead.

  4686   Wed May 11 11:05:45 2011 steveUpdateSUSETMX_OPLEV : He-Ne replaced

Quote:

I found that a He-Ne laser which has been used for ETMX_OPLEV was NOT giving the light.

Since I didn't find the switch key for it I have no idea if the laser is simply off or dead.

 The dead laser was replaced by new JDSU 1103P of  2.6mW. The return beam is big ~5 mm diameter of 0.3 mW,  1400 counts

  4693   Wed May 11 21:58:30 2011 ranaUpdateSUSETMX_OPLEV : He-Ne replaced

Quote:

Quote:

I found that a He-Ne laser which has been used for ETMX_OPLEV was NOT giving the light.

Since I didn't find the switch key for it I have no idea if the laser is simply off or dead.

 The dead laser was replaced by new JDSU 1104P of  2.6mW. The return beam is big ~5 mm diameter of 0.3 mW,  1400 counts

 Whenever replacing any Oplev laser, please also put into the ELOG when it was installed so that we have an electronic record of the laser lifetime.

  14090   Fri Jul 20 07:43:54 2018 SteveSummarySUSETMY

 

 

  8646   Mon May 27 21:38:53 2013 AnnalisaUpdate40m upgradingETMY - Beam Dumps

 

I put many razor dumps along the IR/green path. The rejected beam from the IR Faraday needs to be dumped (about 1.5 mW). I used all the new razor blade I had, so I need one more for that beam.

The IR reflection of the Harmonic separator right after the doubler needs to be dumped in a better way. At the moment there is a black screen, but we need something suitable to dump more than 300 mW.

After the second steering mirror along the green beam path there is a very small transmission (about 6 uW), which is difficult to dump because there is no space enough. Can it be dumped with a black screen?

 

The Oplev has a lot of reflection hitting the central BS (The BS for the transmitted beam). It is very difficult to dump them without intercepting the main beam path. Maybe we have to slightly change the Oplev beam angle to avoid so many reflections.

 

  9451   Mon Dec 9 11:44:08 2013 SteveUpdateendtable upgradeETMY optable got new beam dump

Aluminum shield replaced by razor beam dump.

  4983   Mon Jul 18 15:51:04 2011 steveUpdateSUSETMY spectra of oplev error signal
  5300   Thu Aug 25 08:12:09 2011 steveUpdateSUSETMY & ITMY wiped, replaced, vac door on

Quote:

We've closed up ETMX:

  • the optic was drag wiped
  • the suspension tower was put back in place
  • earthquake stops were backed off the appropriate number of turns, and de-ionized
  • chamber door was put on

 jamie, jenne, kiwamu, suresh, steve

 

ETMY and ITMY were treated the same way as ETMX. The BS chamber was closed with heavy vac door yesterday also. The IOO access connector's inner jamnuts are torqued to 45 ft/lbs as all vac door bolts.

The vac envelope is ready for pumpdown condition, except ITMX chamber with light atm door cover.

Jenne will summeries the condition of dust on the  TMs before and after the drag wipes.

  12961   Mon May 1 17:14:58 2017 SteveUpdateCamerasETMY & MC2 ccd cameras removed

MC2 ccd camera is replaced by Olympus 570 zoom temporarly.

So as the ETMY ccd camera is replaced by Cannon Rebel.

Both viewport are under Lexan protection and covered by Aluminum foil....still, turn all lighting off if you do not want room light in the IFO

 

Do not remove Lexan shield!

 

  8637   Fri May 24 02:12:50 2013 AnnalisaUpdate40m upgradingETMY - Mode Matching for green

 

 Mode Matching calculation for green beam - Yarm

After measuring the beam radius out from the Faraday for the green, I made the calculation to match the green beam mode with the IR mode inside the arm.

The beam waist after the Faraday is elliptical, and I found the following value for the waist:

w0x = 3.55e-5 m @ z0x = -0.042 m

w0y = 2.44e-5 m @ z0y = -0.036 m

(the origin of the z axis is the output of the Faraday, so the waist is inside the Faraday itself)

I did the calculation using a la mode, using as beam waist and its position the following values:

w0 = sqrt(w0x*w0y) = 2.943e-5 m @ z0 = (z0x+z0y)/2 = -0.039 m

The results are shown in the attached plots.

                      Focal length (m)             position (m)

lens1            0.125                                0.1416

lens2            0.100                                0.5225

L                    1.000                                1.5748 (fixed lens used to focus transmitted beam)

 

As the first plot shows, the green beam size on the ETMY is about 6mm. My concern is that it could be too big.

The third plot shows the X and Y section of the beam. It is strongly elliptical, but nevertheless the coupling factor calculated with Koji's formula  gives C=0.936 for the astigmatic beam, and C=0.985 for the non astigmatic beam, so it seems to be still ok.

 

 

  8638   Fri May 24 11:38:00 2013 KojiUpdate40m upgradingETMY - Mode Matching for green

I got confused. Is the mode calculation in the cavity correct?
Are you sure the wavelength in the code is 532nm?

The first plot says "the waist radius at ITMY is 2.15mm". This number is already very close to
the waist size of the cavity mode (2.1mm@ITM, 3.7mm@ETM), but the spot radius at ETMY is 6mm.
They are inconsistent.

 

  8639   Fri May 24 12:50:25 2013 AnnalisaUpdate40m upgradingETMY - Mode Matching for green

Quote:

I got confused. Is the mode calculation in the cavity correct?
Are you sure the wavelength in the code is 532nm?

The first plot says "the waist radius at ITMY is 2.15mm". This number is already very close to
the waist size of the cavity mode (2.1mm@ITM, 3.7mm@ETM), but the spot radius at ETMY is 6mm.
They are inconsistent.

 

 Jenne and I just realized that a la mode has 1064e-9 m as default value. I'll change it and make the calculation again.

  8645   Sat May 25 02:03:48 2013 AnnalisaUpdate40m upgradingETMY - Mode Matching for green - new calculation

 Mode matching calculation for green - Yarm

I did again the mode matching calculation. The previous one was using 1064nm as wavelength, so it was wrong.

The seed beam waist and its position are the same as in elog 8637. The new results are shown in the attached graphs.

I got the following values for focal lengths and positions of the two Mode Matching lenses:

 

                    Focal length (m)             Distance from the Faraday output (m)

lens1            0.125                                                  0.1829

lens2           -0.200                                                  0.4398

L                   1.000                                                  1.4986 (fixed)

The position of the lens L has changed because the path lengh has been slightly  reduced. 

The Coupling factor for he astigmatic beam is C = 0.959 (it is C = 0.9974 if we consider the beam as non astigmatic).

I put the lenses and aligned the beam up to the shutter, which has been moved from its initial position because the beam size on it was too large. 

TO DO

The green beam needs to be aligned and sent into the arm cavity. 

Polarization has to be checked.

Many beams still have to be dumped, both in IR and Green paths. 

 

 

 

 

  8662   Fri May 31 11:15:01 2013 AnnalisaUpdateendtable upgradeETMY - Mode Matching for green - new calculation

 Since the beam waist after the Faraday had changed since the last time I measured it (maybe alignment changed a bit), I made a new mode matching calculation for green. I attached the results.

I'm going to align the beam into the Yarm.

RXA: JPG images deleted - replace with PDF please.

  8652   Tue May 28 22:11:23 2013 AnnalisaUpdate40m upgradingETMY - Yarm shutter cable

 

 For some strange reason the Yarm shutter cable runs up to the POY table, where it is connected to another cable going to the rack. It has to be put off from the table, at least. It would be better to have only one cable going directly to the rack.

I roughly aligned the green into the Yarm and I've seen the green beam flashing on the PSL table, but the mode matching is not so good and I get an higher order mode, so I'm going to fix the mode matching tomorrow.

  8751   Wed Jun 26 00:15:51 2013 AnnalisaUpdateGreen LockingETMY - green locking and beat note setup

[Koji, Annalisa]

Alignment improving

  • The alignment of the green beam injected into the Yarm has been improved, and when the green laser is locked on the cavity in 00 mode we moved from 500 cnts to almost  650 cnts in transmission on the PSL table. Now we have about 70 uW transmitted.
  • Since the beam size on the GRTY PD was too big, I put a lens to focus it better.

Beat note setup

  • The Ybeat RFmon was connected back to the power splitter that Yuta put and Manasa temporarily removed (as described in elog 8666), so that now we are using again the SUM port to monitor the beat signal.

TO DO

  • Align better the beams on the BeatPD, in order to get a stronger beat note
  • Calibrate the ALS screen to tune remotely the laser temperature
  • Find the beat note!

 

 

  8628   Thu May 23 12:02:48 2013 SteveUpdateendtable upgradeETMY - oplev

 Temporary oplev in place. The spot on the qpd is still big. My two lens solution did not work.

I will finalize optical component position of the oplev after the the arm transmitted and green beam optics in place. They have priority.

 

  8655   Thu May 30 11:10:00 2013 SteveUpdateendtable upgradeETMY - oplev

Quote:

 Temporary oplev in place. The spot on the qpd is still big. My two lens solution did not work.

I will finalize optical component position of the oplev after the the arm transmitted and green beam optics in place. They have priority.

 

 Oplev spot size on qpd ~ 1 mm 

PS: I realized it later that the returning beam is going through a lens for TRY. This is a nono.

      This beam path will be relayed again as the TRY, green beam and IP-ang get there place.

  8660   Thu May 30 20:45:53 2013 AnnalisaUpdateendtable upgradeETMY - oplev

Quote:

Quote:

 Temporary oplev in place. The spot on the qpd is still big. My two lens solution did not work.

I will finalize optical component position of the oplev after the the arm transmitted and green beam optics in place. They have priority.

 

 Oplev spot size on qpd ~ 1 mm 

PS: I realized it later that the returning beam is going through a lens for TRY. This is a nono.

      This beam path will be relayed again as the TRY, green beam and IP-ang get there place.

 Oplev is disabled. I removed one of the steering mirrors because it was on the green beam  path.

  8594   Fri May 17 00:32:32 2013 AnnalisaUpdate40m upgradingETMY - progress

[Rana, Annalisa] 

 GREEN

 The alignment for the green has been improved, so that we have much more green power.

The first lens position along the IR path has been changed in way to have the beam waist at the center of the first Faraday. In this way we had about 91% of the input power out from it.

The two cylindrical lenses which were used to correct the ellipticity of the beam have been replaced by a single lens. Its focal length is intermediate between the focal lengths of the two cylindrical. 

Moving the position of the lens before the doubler crystal and improving the alignment we got about 1mW of green light (0.35% of the incoming IR beam).

TO DO

 

After aligning the green beam through the second Faraday, the beam waist of the outgoing beam has to be measured and the mode matching calculation has to be done to choose the two MM lenses. Then the steering mirrors will be placed to send the beam into the arm.

  8597   Fri May 17 18:24:04 2013 AnnalisaUpdate40m upgradingETMY - progress

I aligned the green beam into the Faraday. I needed an HWP to have the right polarization for the light entering the Faraday itself.

I tried to dump as much beams as possible with razor dumps, but eventually I had to use some "temporary solutions" for higher beams, because I didn't find the right mounts for razor dumps.

I measured the beam waist after the Faraday with the beam scan. Analysis and MM calculation to follow.

  8622   Thu May 23 00:16:32 2013 AnnalisaUpdate40m upgradingETMY - progress

 [Annalisa, Koji] 

 GREEN

I aligned back the beam (we lost part of the alignment after we put back the box and after the posts were installed). The green beam out from the crystal is still low, but anyway I get about 1.2 mW of green out from the Faraday. 

TO DO 

Mode Matching calculation (tomorrow)

Fix the dumping situation

Replace some of the mounts with more solid ones (in the future)

TRANSMON PATH

 QPD, PD and Camera have been rotated as Rana suggested last Wednesday. A 1m focal length lens is on the main beam transmitted path (before the harmonic separator), and the beam diameter on the QPD is about 5mm. We put another lens with a shorter focal length to put the PD very close to the beam waist and in way to have a reasonable beam size on the camera. Tomorrow I will write down all the correct sizes of the beams.

OPLEV 

(for Steve) I marked a possible beam path for the Oplev (the laser is not in the right place in the picture, but I left it in the correct place on the table). I also put the QPD for the IP-ANG, so we know in which part of the table the beam can be steered.

The space in the red rectangle (right corner) has to be left empty to put a PD for the rejected beam from the green Faraday.

 

 

 

 

 

 

 

 

 

 

 

 

 

 

  8768   Thu Jun 27 17:41:08 2013 AnnalisaUpdateGreen LockingETMY -beat note found!

 

 Y arm beat note found!

Procedure

  • Arm lock on IR to align the mirrors
  • Green Laser locked on the arm
  • Green Transmission on the PSL and PSL green beam aligned into the BeatPD: they have been aligned both in the near field (looking at the beam on the camera) and in the far field (removing the DC PD and looking at the two beams on the wall)
  • Checked the PSL temperature and, following the plot of the beat note measurement between "Alberto" laser and PSL reported in elog 8396, I got an idea of the range of temperature where the beat note could be found (I used the values of the second curve)
  • Scanned the Y-green laser temperature using the slow servo on the ALS command window

Data

  • PSL temperature = 31.58°C
  • PSL slow servo temperature offset = 0
  • "Alberto" laser temperature = 40.35 °C (ADJ = 0)
  • Thermal output offset on the ALS screen = -11140
  • Beat note frequency = 22 MHz
  • Beat note amplitude = -31.7 dBm

The green transmission on the PSL reads about 500 cts, and the transmitted power is about 50 uW.

(the second peak on the screen in the picture is the 29 MHz of the MC)

  16687   Mon Feb 28 15:51:07 2022 Ian MacMillanUpdateSUSETMY 1Y4 Electronics Replacement

[Paco, Ian]

paco helped me wire the ETMY 1Y4 rack. We wired the following (copied from Koji's email):

  1. Use DB9-DB9 to complete the wiring between
    1. 16bit DAC AI Chassis - End DAC Adapter (4 cables)
    2. End DAC Adapter - HAM-A Coil Driver (2 cables)
    3. AA Chassis - End ADC Adapter (2 cables)
  2. Koji already brought two special DB9-DB15 cables (in plastic bags) to the end. They connect the HAM-A coil drivers to the satellite amp. At this time, we skip Low Noise HV Bias Driver.
  3. Bring two 30ft DB9 (called #1, aka D2100675-01) cables from the office area to the end. I collected one end and left them there.
  4. All the new units have +/-18V DC supply in the back. Find the orange cables behind the 40m vacuum duct around Y-end and connect the units and the DC power strip. Use short cables if possible to save the longer ones.

the cables we used:

Number Used Type of Cable Length
8 DB9 to DB9 2.5 ft
2 DB9 to DB9 5 ft
2 DB9-DB15  
2 DB9 (called #1, aka D2100675-01) 30ft
9 Orange Power Cables ~ 3 ft

I attached pictures below.

  16695   Thu Mar 3 04:11:36 2022 KojiUpdateSUSETMY 1Y4 Electronics Replacement

For the Y-end electronics replacement, we want to remove unused power supplies. In fact, we already removed the +/-5V supplies from the stack. I was checking what supply voltages are used by the Eurocard modules. I found that D990399 QPD whitening board had the possible use of +/-5V.

The 40m Y-end version can be found here D1400415. The +/-5V supply voltages are used at the input stage AD620 and the QPD bias voltage of -5V.

AD620 can work with +/-15V. Also the bias voltage can easily be -15V. So I decided to cut the connector legs and connected +5V line to +15V, and -5V line to -15V.

With this modification, I can say that the eurocards only use the +/-15V voltages and nothing else.

The updated schematics can be found as D1400415-v6

  16696   Thu Mar 3 04:24:23 2022 KojiUpdateSUSETMY 1Y4 Electronics Replacement

The DC power strip at Y-end was connected to the bottom two Sorensen power supplies. They are configured to provide +/-18V.

 

  16699   Thu Mar 3 17:21:11 2022 Ian MacMillanUpdateSUSETMY 1Y4 Electronics Replacement

[Koji, Ian]

1) We attached the 30 coil driving cables to the vacuum feed through to the sat amp  [40m wiki] they run along the cable tray then up and down into the rack.

2) we checked all DB and power cables. We found that the anti-imaging filter had a short and got very hot when plugged in. the back power indicator lights turned on fine but the front panel stayed off. We removed it and replaced it with the one that was on the test stand marked for the BHD. This means we need to fix the broken one and Koji mentioned getting another one.

3) we reassigned the ADC and DAC channels in the iscey model and the asy model. we committed a version before we made any changes.

4) Finally we tested the setup to make sure the ETM was being damped.

Next step:

1) Measure the change of the PD gains and the actuator gains. See pervious elog

  16703   Sat Mar 5 02:03:46 2022 KojiUpdateSUSETMY 1Y4 Electronics Replacement

Oplev saga

Summary

- The new coil driver had not enough alignment range to bring the oplev beam back to the QPD center
- The coil driver output R was reduced from 1.2k to 1.2k//100 = 92.3 +/- 0.4 Ohm
- Now the oplev spot could be moved to the center of the QPD

- The damping gains (POS/PIT/YAW) and the oplev gains were reduced by a factor of 1/10.
- The damping and the oplev servos work now. Fine gain tuning is necessary.

To Do:
- DC value / TF measurements
- Adjust damping gains
- RFM issue
- Connection check
- Cable labeling


== Alignment Range ==

- Since c1auxey was removed, we no longer have C1:SUS-ETMY_PIT_COMM and C1:SUS-ETMY_YAW_COMM. At this moment, all the alignment is taken with the offset input from the fast real-time system via C1:SUS-ETMY_PIT_OFFSET and C1:SUS-ETMY_YAW_OFFSET.

- The oplev spot could not be moved on the center of the QPD without exceeding the DAC output range (~+ or -32000) for the coils. (Attachment 1)

- This is because the old system had a slow but large current range (Rout = 100) and a small current range for the fast control. Until we commission the new HV BIAS Driver, we have to deal with the large DC current with the HAM-A coil driver.

== Modification to the output resistances ==

The following units and the channels were modified. Each channel had a differential current driver and two output resistances of 1.2K. 100Ohm (OHMITE 43F100, 3W) wire wound resistors were added to them in parallel, making the resulting output R of ~92Ohm.

- ETMY HAM A Coil Driver 1: S2100622 (Attachments 2/3) CH1/2/3
- ETMY HAM A Coil Driver 2: S2100621 (Attachments 4/5) CH3

- This modification allowed me to align the oplev spot to the center of the QPD. C1:SUS-ETMY_PIT_OFFSET and C1:SUS-ETMY_YAW_OFFSE are +2725 (8%FS) and -2341 (7%FS), respectively.
- The previous alignment slider values were -0.9392 and 0.7615 (out of 10). These are the reasonable numbers, considering the change of the Rout from 100 to 92Ohm, and the sign flip.
(By the way, autoBurt files for c1auxex and c1auxey were not properly configured and the history of C1:SUS-ETM*_*_COMM was not recorded.)

== Damping Servos ==

- Now, the POS/PIT/YAW servos experience ~x10 gains. So temporarily these gains were reduced (POS 20->2, PIT 6->0.6, YAW 4->0.4) and the loops are stable when engaged.
- Also the gains of the OPLEV servos were reduced from -4.5 to -0.45. The loops are stable when engaged.

== Snapshot of the working condition ==

Attachent 6 shows the screenshot for the snapshot of the working condition.


To Do

- The damping servos were tested without proper PD whitening compensation.
  -> It turned out this is not necessary as our modified PD whitening has the pole and zero at the same freqs as before.

- Compare the DC values of the OSEM outputs and compensate for the gain increase by the "cts2um" filter.

- The end RTS suffers from the RFM issue. There is no data transmitted from the vertex to the end. I suspect we need to restart the c1rfm process. But this will likely suspend all the vertex real-time machines. Careful execution is necessary.

- c1iscey has all the necessary analog connections. But they are not tested. When we lock the green/IR cavity, we'll need them.

- The cable labeling is only half done.

  16738   Mon Mar 21 14:22:52 2022 AnchalUpdateSUSETMY Alignmnet offsets needed to be changed

I'm not sure why but the PIT and YAW offset values of +2725 and -2341 were not sufficient for the reflected OPLEV beam to reach the OPLEV QPD. I had to change the C1:SUS-ETMY_PIT_OFFSET to 5641 and C1:SUS-ETMY_YAW_OFFSET to -4820 to come back to center of the OPLEV QPD. We aligned the Oplevs to center before venting, so hopefully this is our desired ETMY position.

On another note, the issue of ETMY unable to damp was simple. The alignment offsets were on to begin with with values above 1000. This meant that whenever we enabled coil output, ETMY would necessarily get a kick. All I had to do was keep alignment offsets off before starting the damping and slowly increase the alignment offsets to desired position.

Quote:
 

- This modification allowed me to align the oplev spot to the center of the QPD. C1:SUS-ETMY_PIT_OFFSET and C1:SUS-ETMY_YAW_OFFSE are +2725 (8%FS) and -2341 (7%FS), respectively.

  14569   Thu Apr 25 00:30:45 2019 gautamUpdateSUSETMY BR mode

We briefly talked about the bounce and roll modes of the SOS optic at the meeting today. 

Attachment #1: BR modes for ETMY from my free-swinging run on 17 April. The LL coil has a very different behavior from the others.

Attachment #2: BR modes for ETMY from my free-swinging run on 18 April, which had a macroscopically different bias voltage for the PIT/YAW sliders. Here too, the LL coil has a very different behavior from the others.

Attachment #3: BR modes for ETMX from my free-swinging run on 27 Feb. There are many peaks in addition to the prominent ones visible here, compared to ITMY. The OSEM PD noise floor for UR and SIDE is mysteriously x2 lower than for the other 3 OSEMs???

In all three cases, a bounce mode around 16.4 Hz and a roll mode around 24.0 Hz are visible. The ratio between these is not sqrt(2), but is ~1.46, which is ~3% larger. But when I look at the database, I see that in the past, the bounce and roll modes were in fact at close to these frequencies.

In conclusion:

  1. the evidence thus far says that ETMY has 5 resonant modes in the free-swinging data between 0.5 Hz and 25 Hz.
  2. Either two modes are exactly degenerate, or there is a constraint in the system which removes 1 degree of freedom.
  3. How likely is the latter? As any mechanical constraint that removes one degree of freedom would presumably also damp the Qs of the other modes more than what we are seeing.
  4. Can some large piece of debris on the barrel change the PIT/YAW eigenvectors such that the eigenvalues became exactly degenerate?
  5. Furthermore, the AC actuation vectors for PIT and YAW are not close to orthogonal, but are rotated ~45 degrees relative to each other.

Because of my negligence and rushing the closeout procedure, I don't have a great close-out picture of the magnet positions in the face OSEMs, the best I can find is Attachment #4. We tried to replicate the OSEM arrangement (orientation of leads from the OSEM body) from July 2018 as closely as possible.

I will investigate the side coil actuation strength tomorrow, but if anyone can think of more in-air tests we should do, please post your thoughts/poetry here.

  2259   Thu Nov 12 17:24:29 2009 Koji, Joe, PeterConfigurationCDSETMY CDS test started

We started the test of the new CDS system at ETMY.

The plan is as follows:
We do the ETMY test from 9:30 to 15:00 at ETMY from Nov 12~17. This disables the ETMY during this period.
From 15:00 of the each day, we restore the ETMY configuration and confirm the ETMY work properly.


Today we connected megatron to the existing AA/AI modules via designated I/F boxes. The status of the test was already reported by the other entry.

During the test, c1iscey was kept running. We disabled the ETMY actuation by WatchDog. We did not touch the RFM network.

After the test we disconnected our cables and restored the connection to ICS110B and the AI/AA boards.

The WatchDog switches were released.

The lock of the ETMY was confirmed. The full interferometer was aligned one by one. Left in the full configuration with LA=off.

  2181   Thu Nov 5 16:24:59 2009 KojiUpdateCDSETMY CDS test stuff

Joe, Peter, Jay, Koji, Rana

We put the new CDS stuff at Y end 1Y9 rack.

Items

  • megatron
  • wireless router
  • IO chasis (black)
  • Extention cable (between megatron & IO chasis)
  • 1 ADC card
  • 1 DAC card
  • 1 BIO card
  • The adapter box for ADC
  • The adapter box for DAC
  • The adapter box for BIO
  • 2x IDC-DB37 cable for the ADC box - AA chasis
  • 1x IDC cable for the DAC box - Pentek
  • 1x DB cable for the BIO box
  • 1x +/-15V cable for the BIO box
  1321   Wed Feb 18 21:03:22 2009 ranaUpdateCamerasETMY Camera work not elogged!

The control room video is showing us a false ETMY image. Who worked on the ETMY camera or video today??!!

  14075   Tue Jul 17 01:07:40 2018 gautamUpdateSUSETMY EQ stops

For the heater setup on EY table, I EQ-stopped ETMY. Only the face EQ stops (3 on HR face, 2 on AR face) were engaged. The EY Oplev HeNe was also shutdown during this procedure. 

  14630   Wed May 22 11:53:50 2019 gautamUpdateSUSETMY EQ stops backed out

Yesterday we noticed that the POS and SIDE eigenmodes were degenerate (with 1mHz spectral resolution). Moreover, the YAW peak had shifted down by ~500 mHz compared to earlier this week, although there was still good separation between PIT and YAW in the Oplev error signals. Ideas were (i) check if EQ stops were not backed out sufficiently, and (ii) look for any fibers/other constraints in the system. Today morning, I inspected the optic again. I felt the EQ stop viton tips were a bit close to the optic, so I backed them out further. Apart from this, I adjusted the LR and SIDE OSEM position in their respective holders to make the sensor voltages closer to half-light. Kicked the optic again just now, let's see if there is any change.

Remaining tasks:

  1. Check EY table leveling.
  2. Check EY actuation matrix diagonality using this technique.
  3. Check that IR resonances are seen (and all the usual pre-pumpdown alignment checks).
  4. Take close out pictures.
  5. Heavy doors on, pump down.

If everything goes smoothly, I think we should plan for the heavy doors going back on and commencing the pumpdown tomorrow. After discussion with Koji, we came to the conclusion that it isn't necessary to investigate IPANG (high likelihood of it falling off the steering optics during the pumpdown) / AS beam clipping (no strong evidence that this is a problem) for this vent.

Update 1235: Indeed, the eigenmodes are back to their positions from earlier this week. Indeed, the POS and SIDE modes are actually better separated! So, the OSEM/magnet and EQstop/optic interactions are non-negligible in the analysis of the dynamics of the pendulum.

  14629   Tue May 21 21:33:27 2019 gautamUpdateSUSETMY HR face cleaned

[koji, gautam]

We executed this plan. Photos are here. Summary:

  1. Optic was EQ-stopped (face stops only)., with the OSEMs in situ. We tried to do this as evenly as possible to avoid any magnets getting stuck on OSEMs.
  2. We used the specially procured acetone from Chub to drag wipe the HR face. This was a definite improvement, we should always get the correct grade of solvents when we attempt cleaning optics.
  3. It was observed that drag-wiping did not really have the desired cleaning effect. So Koji went in with hemostat / lens tissue soaked in acetone and wiped the HR face. This improved the situation.
  4. Applied a layer of F.C. Waited for it to dry, and then peeled it off. Under the green flashlight, the optic still looks horrific - but we decided against further drag-wiping/first-contacting. If the loss is truly 50 ppm, this is totally not a show-stopper for now.
  5. Suspension cage was replaced. EQ stops were released. Bias voltages were adjusted to bring the Oplev spot back to the center of the QPD. Now a free-swinging data collection is ongoing...
The following optics were kicked:
ETMY
Tue May 21 22:58:18 PDT 2019
1242539916

So if nothing, we got to practise this new wiping technique with OSEMs in situ successfully.

Quote:
 
  1. Prepare ETMY for first contact cleaning to remove the residual piece. 
    • Drag wipe the HR surface with dehydrated acetone 
    • Apply F.C. as usual, inspect the HR face after peeling for improvement if any.
    • This will give us a chance to practise the F.C.ing with the optic EQ-stopped (moving cage etc).
  4057   Wed Dec 15 13:36:44 2010 josephbUpdateCDSETMY IO chassis update

I gave Alex a sob story over lunch about having to go and try to resurrect dead VME crates.  He and Rolf then took pity on me and handed me their last host interface board from their test stand, although I was warned by Rolf that this one (the latest generation board from One Stop) seems to be flakier than previous versions, and may require reboots if it starts in a bad state.

Anyways, with this in hand I'm hoping to get c1iscey damping by tomorrow at the latest.

  4058   Wed Dec 15 14:23:32 2010 KojiUpdateCDSETMY IO chassis update

Great!

I wish this board works fine at least for several days...

Quote:

I gave Alex a sob story over lunch about having to go and try to resurrect dead VME crates.  He and Rolf then took pity on me and handed me their last host interface board from their test stand, although I was warned by Rolf that this one (the latest generation board from One Stop) seems to be flakier than previous versions, and may require reboots if it starts in a bad state.

Anyways, with this in hand I'm hoping to get c1iscey damping by tomorrow at the latest.

 

  14625   Mon May 20 17:12:57 2019 gautamUpdateSUSETMY LL adjustment

Following the observation that the response in the LL shadow sensor was lower than that of the others, I decided to pull it out a little to move the signal level with nominal DC bias voltage applied was closer to half the open-voltage. I also chose to rotate the SIDE OSEM by ~20 degrees CCW in its holder (viewed from the south side of the EY chamber), to match more closely its position from a photo prior to the haphazhard vent of the summer of 2018. For the SIDE OSEM, the theoretical "best" alignment in order to be insensitive to POS motion is the shadow sensor beam being horizontal - but without some shimming of the OSEM in the holder, I can't get the magnet clear of the teflon inside the OSEM.

While I was inside the chamber, I attempted to minimize the Bounce/Roll mode coupling to the LL and SIDE OSEM channels, by rotating the Coil inside the holder while keeping the shadow sensor voltage at half-light. To monitor the coupling "live", I set up DTT with 0.3 Hz bandwidth and 3 exponentially weighted averages. For the LL coil, I went through pi radians of rotation either side of the equilibrium, but saw no significant change in the coupling - I don't understand why.

In any case, this wasn't the most important objective so I pushed ahead with recovering half-light levels for all the shadow sensors and closed up with the light doors. I kicked the optic again at 1712:14 PDT, let's see what the matrix looks like now.


before starting this work, i had to key the unresponsive c1auxey VME crate.

  12349   Thu Jul 28 17:48:17 2016 gautamUpdateGeneralETMY LR magnet broke off

While ETMX magnets were curing, I wanted to try and suspend ETMY in the endchamber, put in the OSEMS and see if the magnets aligned well with the coils, and run the same type of diagnostics we have been doing for ETMX. However, while I was trying to slip the optic into the wire, the UL magnet on ETMY broke off. I recovered the magnet and now both optic and magnet are back in the cleanroom. The magnet dumbbell has been cleaned with acetone and then sandpaper to remove residual epoxy - it remains to clean the residue off the optic itself before re-gluing the magnet tonight

I also noticed that the existing wire in the suspension had a kink in it. It looks fairly sharp, and I think we should change the wire while re-inserting the optic. Putting the optic into an existing loop of wire is tricky, as if you go in from the front of the suspension cage, the magnets on the AR side attract the wire, and makes it quite difficult to loop the wire around. I have to think of some way of holding the wires in place while the optic is being placed, and then, once the optic is roughly in position, slip the wire into the grooves in the standoffs. 


I took the opportunity to replace the face OSEM coil holder screws while the chamber was open. 

EDIT 9 August 2016: It was in fact the LR magnet that was knocked off.

  14447   Mon Feb 11 16:38:34 2019 gautamUpdateLSCETMY OL calibration updated

Since we changed the HeNe, I updated the calibration factors, and accepted the changes in the SDF.

DOF OLD [urad/ct] NEW [urad/ct]
PITCH 140 176
YAW 143

193