rob | Configuration | SUS | watchdog rampdown |
I've changed the watchdog rampdown script so it brings the SUS watchdogs to 220, instead of the 150 it previously targeted. This is to make tripping less likely with the jackhammering going on next door. I've also turned off all the oplev damping. |
11487
|
Mon Aug 10 15:25:05 2015 |
Steve | Update | PEM | wasp nest |
The wasp nest will be removed tomorrow from from the out side of the east arm window.
The resonant frequency of the newly arrived gravity bee detector is not known.
|
15566
|
Wed Sep 9 20:52:45 2020 |
rana | Summary | IOO | wandering line in IMC |
since the summary pages are working again, I was clicking through and noticed that there's a wandering peak in the whitened IMC spectrogram that goes from 10-30 Hz over the course of a day.
https://nodus.ligo.caltech.edu:30889/detcharsummary/day/20200909/ioo/
anyone know what this is ? |
2823
|
Wed Apr 21 10:09:23 2010 |
kiwamu | Update | Green Locking | waist positon of Gaussian beam in PPKTP crystals |
Theoretically the waist position of a Gaussian beam (1064) in our PPKTP crystal differs by ~6.7 mm from that of the incident Gaussian beam.
So far I have neglected such position change of the beam waist in optical layouts because it is tiny compared with the entire optical path.
But from the point of view of practical experiments, it is better to think about it.
In fact the result suggests the rough positioning of our PPKTP crystals;
we should put our PPKTP crystal so that the center of the crystal is 6.7 mm far from the waist of a Gaussian beam in free space.
(How to)
The calculation is very very simple.
The waist position of a Gaussian beam propagating in a dielectric material should change by a factor of n, where n is the refractive index of the material.
In our case, PPKTP has n=1.8, so that the waist position from the surface of the crystal becomes longer by n.
Now remember the fact that the maximum conversion efficiency can be achieved if the waist locates at exact center of a crystal.
Therefore the waist position in the crystal should be satisfied this relation; z*n=15 mm, where z is the waist position of the incident beam from the surface and 15 mm is half length of our crystal.
Then we can find z must be ~8.3 mm, which is 6.7 mm shorter than the position in crystal.
The attached figure shows the relation clearly. Note that the waist radius doesn't change. |
2850
|
Tue Apr 27 14:18:53 2010 |
kiwamu | Update | Green Locking | waist positon of Gaussian beam in PPKTP crystals |
The mode profile of Gaussian beams in our PPKTP crystals was calculated.
I confirmed that the Rayleigh range of the incoming beam (1064 nm) and that of the outgoing beam (532 nm) is the same.
And it turned out that the waist postion for the incoming beam and the outgoing beam should be different by 13.4 mm toward the direction of propagation.
These facts will help us making optical layouts precisely for our green locking.
(detail)
The result is shown in the attached figure, which is essentially the same as the previous one (see the entry).
The horizontal axis is the length of the propagation direction, the vertical axis is the waist size of Gaussian beams.
Here I put x=0 as the entering surface of the crystal, and x=30 mm as the other surface.
The red and green solid curve represent the incoming beam and the outgoing beam respectively. They are supposed to propagate in free space.
And the dashed curve represents the beams inside the crystal.
A trick in this calculation is that: we can assume that the waist size of 532 nm is equal to that of 1064 nm divided by sqrt(2) .
If you want to know about this treatment in detail, you can find some descriptions in this paper;
"Third-harmonic generation by use of focused Gaussian beams in an optical super lattice" J.Opt.Soc.Am.B 20,360 (2003)" |
3325
|
Thu Jul 29 21:13:39 2010 |
Dmass | Update | Green Locking | waist positon of Gaussian beam in PPKTP crystals |
Quote: |
The mode profile of Gaussian beams in our PPKTP crystals was calculated.
I confirmed that the Rayleigh range of the incoming beam (1064 nm) and that of the outgoing beam (532 nm) is the same.
And it turned out that the waist postion for the incoming beam and the outgoing beam should be different by 13.4 mm toward the direction of propagation.
These facts will help us making optical layouts precisely for our green locking.
(detail)
The result is shown in the attached figure, which is essentially the same as the previous one (see the entry).
The horizontal axis is the length of the propagation direction, the vertical axis is the waist size of Gaussian beams.
Here I put x=0 as the entering surface of the crystal, and x=30 mm as the other surface.
The red and green solid curve represent the incoming beam and the outgoing beam respectively. They are supposed to propagate in free space.
And the dashed curve represents the beams inside the crystal.
A trick in this calculation is that: we can assume that the waist size of 532 nm is equal to that of 1064 nm divided by sqrt(2) .
If you want to know about this treatment in detail, you can find some descriptions in this paper;
"Third-harmonic generation by use of focused Gaussian beams in an optical super lattice" J.Opt.Soc.Am.B 20,360 (2003)"
|
If I understand your elog, you are just calculating the the offset in position space that you get by having a refractive index.
Did you end up changing the mode matching so that the rayleigh range (which changes with refractive index) was confocally focused inside the crystal (e.g. Zr = 15 mm?
|
3327
|
Thu Jul 29 22:58:25 2010 |
kiwamu | Update | Green Locking | waist positon of Gaussian beam in PPKTP crystals |
- As you said, I just calculated the waist position in the crystal because the speed of light changes in a medium and eventually the waist position also changes.
- Yes, I did. Once you get a beam with the right waist size, you just put your crystal at the waist position with the offset.
In fact you don't have to think about the rayleigh range inside of the crystal because what we care is the waist size and it doesn't change.
Quote: |
If I understand your elog, you are just calculating the the offset in position space that you get by having a refractive index.
Did you end up changing the mode matching so that the rayleigh range (which changes with refractive index) was confocally focused inside the crystal (e.g. Zr = 15 mm?
|
|
3328
|
Fri Jul 30 00:02:15 2010 |
Dmass | Update | Green Locking | waist positon of Gaussian beam in PPKTP crystals |
Quote: |
- As you said, I just calculated the waist position in the crystal because the speed of light changes in a medium and eventually the waist position also changes.
- Yes, I did. Once you get a beam with the right waist size, you just put your crystal at the waist position with the offset.
In fact you don't have to think about the rayleigh range inside of the crystal because what we care is the waist size and it doesn't change.
Quote: |
If I understand your elog, you are just calculating the the offset in position space that you get by having a refractive index.
Did you end up changing the mode matching so that the rayleigh range (which changes with refractive index) was confocally focused inside the crystal (e.g. Zr = 15 mm?
|
|
I thought we cared about satisfying the confocal focusing parameter, that is to say we want to set Zr = 2L_crystal. If Zr changes inside the crystal, this is the number we care about..isn't it NOT the waist size, but the rayleigh range we care about? I am not entirely sure what youre response is saying you did...
- Calculate Zr = pi * wo^2/(lamba/n)
- Do mode matching to get this wo in free space
- Calculate the offset you need to move the oven by using n
- Move hte ovens
OR
- Calculate Zr = pi*wo^2/(lamba)
- Do mode matching to get this in free space
- Calculate the offset you need to move your ovens using n
- Move your ovens
I guess the waist size would also let me know - are you using 69 um or 53 um waist size? |
6305
|
Wed Feb 22 16:55:16 2012 |
Jamie | Update | SUS | wacky state of SUS input matrices |
While Kiwamu and I were trying to investigate the the vertex glitches we were noticing excess noise in ITMX, which Kiwamu blamed on some sort of bad diagonalization. Sure enough, the ITMX input matrix is in the default state [0], not a properly diagonalized state. Looking through the rest of the suspensions, I found PRM also in the default state, not diagonalized.
We should do another round of suspension diagonalization.
Kiwamu (or whoever is here last tonight): please run the free-swing/kick script (/opt/rtcds/caltech/c1/scripts/SUS/freeswing) before you leave, and I'll check the matrices and update the suspensions tomorrow morning.
[0]
0.25 |
0.25 |
0.25 |
0.25 |
0 |
1.66 |
1.66 |
-1.66 |
1.66 |
0 |
1.66 |
-1.66 |
-1.66 |
1.66 |
0 |
0 |
0 |
0 |
0 |
1 |
|
6307
|
Thu Feb 23 02:20:07 2012 |
Zach | Update | SUS | wacky state of SUS input matrices |
This reminds me that the whole Dr. SUS situation never got taken care of. Where I left off, I was having issues pulling 40m data with NDS2 (which is what all the diagonalization scripts use).
What is the deal with 40m+NDS2? If it is till no-go, can we have a consensus on whether this is too important to wait for? If so, I will rewrite the scripts to use NDS and we can upgrade to NDS2 once we can prove we know how to use it.
Quote: |
While Kiwamu and I were trying to investigate the the vertex glitches we were noticing excess noise in ITMX, which Kiwamu blamed on some sort of bad diagonalization. Sure enough, the ITMX input matrix is in the default state [0], not a properly diagonalized state. Looking through the rest of the suspensions, I found PRM also in the default state, not diagonalized.
We should do another round of suspension diagonalization.
Kiwamu (or whoever is here last tonight): please run the free-swing/kick script (/opt/rtcds/caltech/c1/scripts/SUS/freeswing) before you leave, and I'll check the matrices and update the suspensions tomorrow morning.
[0]
0.25 |
0.25 |
0.25 |
0.25 |
0 |
1.66 |
1.66 |
-1.66 |
1.66 |
0 |
1.66 |
-1.66 |
-1.66 |
1.66 |
0 |
0 |
0 |
0 |
0 |
1 |
|
|
12781
|
Tue Jan 31 22:15:02 2017 |
Johannes | Update | CDS | vme crate backplane adapter boards |
I made a crude sketch for how Lydia and I envision the connector situation on the back of the vme crates to be solved. Essentially the side panels of each crate extend about 2" (52 mm) beyond the edge of the DIN connectors. This is plenty of space for a simple PCB board. The connector of choice is D-Sub. We can split the 64 used pins into 2x 37 D-Sub OR (2x25 pin + 1x15pin). The former has fewer cables, but a few excess unused leads. A quick google search showed me that it is much cheaper to get twisted pair cables for 15 and 25 pin D-Subs. From what I remember, the used pins on the DIN connectors are concentrated on the low numbers end and the high numbers end, so might not need the 'middle' connector in many cases if we decide to break it up into three. I have to check this with Lydia though.
The D-Sub connectors would be panel mounted, for which we need a narrow panel piece with dsub cutouts. We can run horizontal struts across the vme crate from side panel to side panel. This way the force upon cable (dis)connection is mostly on the panel which is attached to the struts which are attached to the crate. This will also prevent gravitational sag or cable strain from pulling on the DIN connection, and we can use twisted pair cables with backshell, screws, and strain reliefs.
I was lookng into getting started with the PCB when Altium complained that the license is expired and to renew it. This is a relatively simple board layout so some free software out there is probably enough. |
12810
|
Tue Feb 7 19:14:59 2017 |
Johannes | Update | CDS | vme crate backplane adapter board layout |
After fighting with Altium for what seems like an eternity I have finished putting my vision of the vme crate backplane adapter board into an electronic format. It is dimensioned to fill the back space of the crate exactly. The connectors are panel mount and the PCB attaches to the connectors with screws, such that the whole thing will be mechanically much more stable than the current configuration. A mounting bracket will attach to horizontal struts that need to be installed in the crates, mechanical drawings to follow. |
5434
|
Fri Sep 16 16:07:28 2011 |
steve | Update | SAFETY | visitors safety training |
Paul, Mirko and Katrin visiting grad students received the 40m basic safety training. |
3682
|
Fri Oct 8 17:36:16 2010 |
steve | Frogs | Photos | visiting undergrads |
Prof Alan Weistein guided the 24 student through the 40m. His performance was rated as an enthusiastic 9.5 |
4447
|
Mon Mar 28 16:19:23 2011 |
steve | Frogs | Photos | visithing 5th graders |
Suresh is captivating his audience with gravity waves on last Friday, March 25 |
3875
|
Sat Nov 6 01:54:15 2010 |
Frank | Summary | Computers | virus definition file update on laptop for dinocam |
i took some pictures with the dinocam this afternoon. I used the laptop computer next to it using wireless lan connection to the caltech network to send the pictures to me.
The installed anti virus software was bitching about the old database and wanted to update that. As the installed virus definition database was from mid last year i agreed and started downloading the update. As the file was huge (~100MB) it wasn't finished when i left. computer is still running and probably waiting for instructions.
Will come back on the weekend to finalize the new virus definition file database installation. |
1611
|
Wed May 20 01:53:48 2009 |
rob, pete | Update | Locking | violin mode filters in drstep_bang |
Recently the watch script was having difficulty grabbing a lock for more than a few seconds. Rob discovered that the violin notch filters which were activated in the script were causing the instability. We're not sure why yet. The script seems significantly more stable with that step commented out. |
8612
|
Wed May 22 00:42:13 2013 |
Koji | Summary | SUS | violin Q |
While looking at the decay of the violin mode of the PRM, I made a simple measurement of the decay rate.
Error signal: REFL33I
The peak @628Hz became 0.372 to 0.303 in 60 sec.
-> Half life of the amplitude T_{1/2} is 203sec.
Q = 4.53 f0 T_{1/2} = 5.8 x10^5 |
8036
|
Fri Feb 8 12:43:26 2013 |
yuta | Update | Computers | videocapture.py now supports movie capturing |
I updated /opt/rtcds/caltech/c1/scripts/general/videoscripts.py so that it supports movie capturing. It saves captured images (bmp) and movies (mp4) in /users/sensoray/SensorayCaptures/ directory.
I also updated /opt/rtcds/caltech/c1/scripts/pylibs/pyndslib.py because /usr/bin/lalapps_tconvert is not working and now /usr/bin/tconvert works.
However, tconvert doesn't run on ottavia, so I need Jamie to fix it.
videocapture.py -h:
Usage:
videocapture.py [cameraname] [options]
Example usage:
videocapture.py MC2F -s 320x240 -t off
(Camptures image of MC2F with the size of 320x240, without timestamp on the image. MUST RUN ON PIANOSA!)
videocapture.py AS -m 10
(Camptures 10 sec movie of AS with the size of 720x480. MUST RUN ON PIANOSA!)
Options:
-h, --help show this help message and exit
-s SIZE specify image size [default: 720x480]
-t TIMESTAMP_ONOFF timestamp on or off [default: on]
-m MOVLENGTH specity movie length (in sec; takes movie if specified) [default: 0] |
2120
|
Mon Oct 19 18:14:28 2009 |
rob | Update | Cameras | video switch broken |
The Chameleon HB (by Knox) video switch that we use for routing video signals into the control room monitors is broken. Well, either it's broken, or something is wrong with the mv162 EPICS IOC which communicates with it via RS-232. Multiple reboots/resets of both machines has not yet worked. The CHHB has two RS-232 inputs--I switched to the second one, and there is now one signal coming through to a monitor but no switching yet. I've been unable to further debug it because we don't have anything in the lab (other than the omega iserver formerly used for the RGA logger) which can communicate with RS-232 ports. I've been trying to get this thing (the iserver) working again, but can't communicate with it yet. For now I'm just going to bypass the video switch entirely and use up all the BNC barrel connectors in the lab, so we can at least have the useful video displays back. |
10259
|
Wed Jul 23 10:39:18 2014 |
Steve | Update | Cameras | video quad processors replaced |
Quad processor 2 & 3 were replaced. |
7236
|
Mon Aug 20 18:10:44 2012 |
Jenne | Update | Cameras | video capture script copied over to real scripts directory |
The videocapture.py script is now in ...../scripts/general/ , along with the videoswitch.
Also, there's a button gui on the VIDEO medm screen to capture different camera views. |
7215
|
Fri Aug 17 08:33:46 2012 |
Steve | Update | Cameras | video cameras in the dark |
Quote: |
I optimized the TM views with illuminator light on quad1 It actually looks better there.
I'll post a dark- OSEM light only in jpg tomorrow. ETMY camera is malfunctioning in dark condition now.
|
ALL illuminator lighting are off. ITMX and ETMY looks back lighted. I will check on their apertures.
In order to focus on 1064 resonant spots I tried to restore and align the arms by script. I only got flashes. |
7216
|
Fri Aug 17 09:34:27 2012 |
Koji | Update | Cameras | video cameras in the dark |
I used the LED illuminations at ETMX and BS yesterday for a tour.
I am afraid that I left them on. |
7217
|
Fri Aug 17 10:38:15 2012 |
Steve | Update | Cameras | video cameras in the dark |
> I used the LED illuminations at ETMX and BS yesterday for a tour.
> I am afraid that I left them on.
It was turned off before the picture was taken.
All LED illuminations were turned off. I checked them a few times. |
7219
|
Fri Aug 17 14:45:51 2012 |
rana | Update | Cameras | video cameras in the dark |
The problem with the glow on the ETMY face is due to the red light being scattered off of the optical table from the HeNe laser for the OL. Why is the red light hitting the table?
One way to fix the problem for the camera image is to insert a long pass filter (if Steve can find one).
Edmund Optics: NT62-874
Edmund Optics: NT65-731
Edmund Optics: NT32-759 |
7232
|
Mon Aug 20 09:49:01 2012 |
Steve | Update | Cameras | video cameras in the DARK |
Quote: |
The problem with the glow on the ETMY face is due to the red light being scattered off of the optical table from the HeNe laser for the OL. Why is the red light hitting the table?
One way to fix the problem for the camera image is to insert a long pass filter (if Steve can find one).
Edmund Optics: NT62-874
Edmund Optics: NT65-731
Edmund Optics: NT32-759
|
Atm1, condition: all oplev lasers are off or blocked, green shutters are closed at the ends, PSL out put shutter is closed, all outside LED illuminating are off, all room lights are off
Only the OSEMs are on. ETMY and ITMX are still look like illuminated.
Atm2, condition: open PSL shutter. ETMY at 11 o'clock and ETMX 1 o'clock bright scattered spot of 1064 nm are visible
Atm3, condition: closed PSL shutter and restored all oplev He/Ne lasers, it is visible at ETMY
Next: I will disconnect power to OSEMs at ETMY |
3347
|
Mon Aug 2 10:26:15 2010 |
steve | Configuration | VAC | very slow pumpdown completed |
Quote: |
Bob and Steve closed BS chamber with the help of the manual Genie lift and the pump down started. The PSL shutter was closed and manual block was placed in the beam path. High voltage power supplies were checked to be off.
Pumping speed ~ 1 Torr/min was achieved at 1/8 of a turn opened roughing valve RV1
|
3 days pump down #69 completed. Thanks to Koji and Kiwamu who did more roughing on Saturday and Sunday. |
3699
|
Tue Oct 12 17:42:57 2010 |
yuta | Update | SUS | very first measurement of Q-values for MC1 |
Background:
Data aquisition system is fixed, and now we can use the Dataviewer to measure Q-values of the ringdowns for each DOF, each optics.
First of all, I measured MC1 suspention damping servo for a test.
What I did:
1. Used DAQ channels activated in this entry(#3690) to see and compare the ringdowns when the damping servo is on and off with the Dataviewer.
2. Plotted the data and fitted the ringdown using this formula;
p[0]*exp(-p[1]*t)*sin(p[2]*t+p[3])+p[4]
I used python's scipy.optimize.leastsq for the fitting.
3. Calculated the resonant frequency f0 and Q-value using following formulas;
f0=2*pi*sqrt(p[1]**2+p[2]**2)
Q=f0/(2*pi)/(2*p[1])
4. For plotting, I subtracted the offset(=p[4]).
All parameters I used for this measurement are automatically saved here;
/cvs/cds/caltech/burt/autoburt/snapshots/2010/Oct/12/13:07/c1mcs.epics
(-1,0,1 for all matrix elements, GAIN=3,3,3,150 for POS,PIT,YAW,SIDE)
Result:
Attached is the plot of each 4 DOF ringdown when servo is off and on.
"servo off" means off for that DOF. Servo for the other 3 DOFs are on.
As you can see clearly, the damping servo is working.
The resonant frequencies and Q-values calculated from the fitting are as follows;
|
servo off |
servo on |
f0 (Hz) |
Q |
f0 (Hz) |
Q |
POS |
0.97 |
large |
0.97 |
16 |
PIT |
0.71 |
96 |
0.73 |
6.9 |
YAW |
0.80 |
100 |
0.82 |
8.9 |
SIDE |
0.99 |
large |
0.99 |
27 |
Resonant frequencies and Q-values have about 1% and 10% error respectively.
I estimated it from my 2-time measurement of the POS ringdown.
Next work:
- Find and modify some scripts to optimize the matrix elements
- Calibrate the displacement
- Do the same thing for other optics
|
3492
|
Tue Aug 31 02:50:45 2010 |
kiwamu | Update | CDS | vertex suspensions controlled by the new CDS |
I plugged the new CDS to the vertex suspensions.
Now PRM, SRM, ITMs, MCs and BS are under the control of the new CDS.
From now on we will never go back to the old system.
Though, the watchdogs are still running on the old system.
So if you need to turn on/off the watchdogs, you can simply enable/disable them from the usual medm screens. |
4165
|
Tue Jan 18 10:25:23 2011 |
steve | Update | PEM | vertex crane work this week |
The crane people are here. The Vertex chambers are covered with plastic. The PSL HEPAs will be running on high during the day.
It is going to be disturbingly noisy and dirty during the day. Please try start working in the evening if it can not wait till next week. |
3624
|
Thu Sep 30 09:34:58 2010 |
steve | Update | General | vertex crane trolly drive fixed |
Quote: |
The vertex crane drive is overheating, it stopped functioning. Service man will be here tomorrow morning.
I crane was just turned on for for may be about 5 minutes. The vertical drive was fine for a while, but the horizontal did not worked at all.
The crane is tagged out again and the controller box is cooling down.
|
Atm1, Vertex crane controller unit on horizontal I-beam.
Module CCES-407 1HP 2.3A 460V/3 phase on the right was replaced. Speed was increased to 30 Hz
Atm2, See burned insulation on black wire that was shorting to the large suppressor resistor on Atm3 One can see the pit marks on the right end of it.
This large power dissipater resistor got hot as the malfunctioning controller 407 broke down. Touching black power wire insulation melted and made a short.
So the heat was created and finally the fuses were blown. The unnecessary large fuses were replaced by small ones.
Fortunately we had one spare controller on hand that made this repair to be done fast.
We used the new Crane Safety Check list from LIGO-E1000379-v1 the first time yesterday before doors were removed. |
9173
|
Mon Sep 30 08:53:22 2013 |
Steve | Update | VAC | vertex crane repair schedule has changed |
Quote: |
Quote: |
The folding beam removed as shown. Two man supporting it while I hammering it out. Pin was dry and it gulled into supporting hinges.
The rotating hinge will be machined and bushing will be added with Zerk fitting or similar. This will allow lubrication in the future.
see elog #9111
|
Atm1, The folding arm is back on with 0.1" misalignment at no load in the trolly's way. The other side of the I beam is 0.02" higher than the main beam.. New bushing and pin were greased up with Krytox before installation.
The axial Zerk 1/8" pipe in the pin upper end can not take any fitting. There is no room. It is taped off.
This gap comes down to ~ 1/16" at fully extended arm with 225 lbs load at the end of it.
The present plan is to grind down the the misalignment of 0.1" for a slow-loaded trolly.
Steve Baker of Konecranes will be back to grind down this ridge and load test at 500 lbs on Tuesday, OCT 1, 2013
|
Konecranes rescheduled the completion of the Vertex crane to Wednesday, Oct. 2 |
9117
|
Mon Sep 9 15:33:06 2013 |
Steve | Update | VAC | vertex crane repair is scheduled |
Quote: |
[Fred Goldbar, Mike Gerfen, Dennis Coyne and Steve]
We inspected the hinge, 1.25" cross pin and I-beams. It is hard to explain what is causing the folding I-beam corner to jam against the main I-beam.
To limit the motion of the folding I-beam cross pin bushing will be added. This will take a week to complete.
|
KoneCrane contact John McDaniel (562) 903 - 1371,
Wednesday, September 18, folding I-beam will be removed. KoneCrane will start working at 7:30am and they should be out by 12:30pm
Friday, September 20, reinstalling machined hinge on the I-beam. Same timing schedule as Wednesday. |
9372
|
Wed Nov 13 08:46:03 2013 |
Steve | Update | VAC | vertex crane repair completed |
nullQuote: |
The smoke alarms were turned off and surrounding areas were covered with plastic.
The folding I-beam was ground down to be in level with the main beam.
Load bearing cable moved into correct position. New folding spring installed.
Crane calibration was done at 500 lbs at the end of the fully extended jib.
Than we realized that the rotating wheel limit switch stopped working.
This means that the crane is still out of order. 
|
New limit switch installed and tested. The crane is back in full operational mode. Two spare limit switches on hand. |
9185
|
Wed Oct 2 14:25:39 2013 |
Steve | Update | VAC | vertex crane is still out of order |
The smoke alarms were turned off and surrounding areas were covered with plastic.
The folding I-beam was ground down to be in level with the main beam.
Load bearing cable moved into correct position. New folding spring installed.
Crane calibration was done at 500 lbs at the end of the fully extended jib.
Than we realized that the rotating wheel limit switch stopped working.
This means that the crane is still out of order. 
|
9111
|
Thu Sep 5 10:47:58 2013 |
Steve | Update | VAC | vertex crane folding issue |
[Fred Goldbar, Mike Gerfen, Dennis Coyne and Steve]
We inspected the hinge, 1.25" cross pin and I-beams. It is hard to explain what is causing the folding I-beam corner to jam against the main I-beam.
To limit the motion of the folding I-beam cross pin bushing will be added. This will take a week to complete.
|
9139
|
Wed Sep 18 14:57:12 2013 |
Steve | Update | VAC | vertex crane folding arm removed |
The folding beam removed as shown. Two man supporting it while I hammering it out. Pin was dry and it gulled into supporting hinges.
The rotating hinge will be machined and bushing will be added with Zerk fitting or similar. This will allow lubrication in the future.
see elog #9111 |
9146
|
Fri Sep 20 16:37:30 2013 |
Steve | Update | VAC | vertex crane folding arm is back on |
Quote: |
The folding beam removed as shown. Two man supporting it while I hammering it out. Pin was dry and it gulled into supporting hinges.
The rotating hinge will be machined and bushing will be added with Zerk fitting or similar. This will allow lubrication in the future.
see elog #9111
|
Atm1, The folding arm is back on with 0.1" misalignment at no load in the trolly's way. The other side of the I beam is 0.02" higher than the main beam.. New bushing and pin were greased up with Krytox before installation.
The axial Zerk 1/8" pipe in the pin upper end can not take any fitting. There is no room. It is taped off.
This gap comes down to ~ 1/16" at fully extended arm with 225 lbs load at the end of it.
The present plan is to grind down the the misalignment of 0.1" for a slow-loaded trolly.
Steve Baker of Konacranes will be back to grind down this ridge and load test at 500 lbs on Tuesday, OCT 1, 2013
|
3616
|
Tue Sep 28 14:12:14 2010 |
steve | Update | General | vertex crane drive is out of order |
The vertex crane drive is overheating, it stopped functioning. Service man will be here tomorrow morning.
I crane was just turned on for for may be about 5 minutes. The vertical drive was fine for a while, but the horizontal did not worked at all.
The crane is tagged out again and the controller box is cooling down. |
2631
|
Tue Feb 23 13:37:04 2010 |
kiwamu and steve | Configuration | VAC | venting the 40m vac envelope |
Kiwamu and Steve have started venting the 40m vacuum envelope.
Preparation:
centered oplevs at resonating cavities,
ITM references were set by green pointer from the ends by Koji,
closed PSL shutter and placed manual block into beam path,
checked jamnuts in locked positions on bellows,
turned HV off at PZT-Jena "steering mirror" power supply and OMC HV ps
checked particle counts,
switched oplev servos off,
set up N2 cylinder to start vent from 1e-6 Torr to 25 Torr,
have ~ 6 cylinders of instrument grade compressed air to bring envelope from 25 Torr to 760 Torr
All three cranes were wiped off today.
|
66
|
Tue Nov 6 09:45:22 2007 |
steve | Summary | SUS | vent sus trend |
The mc optics dragwippings were done by locking optics by eq stops and rotating-moving
cages so access were good. This technic worked well with mc1 & mc2
MC3 osems were reoriented only. |
7560
|
Tue Oct 16 17:13:23 2012 |
Czarina | Summary | General | vent stuff - 4 paths |
I see 4+ possible paths for us to take, in terms of a possible vent in the next few weeks:
No Vent - Just do FPPRMI, using AS55
Mini Vent - Fix REFL path, nothing else. ~1 day at atmosphere
Medium Vent - Fix REFL path, swap G&H mirrors for LaserOptik mirrors (so also resuspend passive TTs, maybe add pitch adjustment option). ~1 week or so at atmosphere - do this rather than Mini if Jan's Finesse calc says the G&H mirrors are too rough
Mega Vent - Fix all the things, do all the things. Long time at atmosphere
The "+" is to take into account all the possible variations on "medium vent". The No, Mini and Medium options assume we'll do the Mega option later, just not immediately. |
7558
|
Tue Oct 16 16:46:20 2012 |
Vent Czar and Czarina | Summary | General | vent stuff |
We're discussing the plan for the next vent. Now that we have all the tip-tilt stuff in hand, we should get ready for the big TT installation vent ASAP. The question is what remains to be done, and when can we be ready
Stuff that needs doing:
* characterize TTs, check electrical connection of quadrupus cables, bake cables and base plates - a week or so (not including baking)
* phase map of LaserOptik mirrors (decide on needed resolution) - a week or so?
* get ready to swap G&H mirrors with LaserOptik mirrors (PR2, PR3, SR2, SR3)
* assembly of black glass beam dumps - one afternoon
* green periscope moving - day or so to make sure greens are resonant in arms, so we have something to align to
* redraw of REFL path?? (rotate 2nd refl mirror to reflect beam to the east, then add mirror to get beam out of IOO chamber viewport), cut hole in BS oplev table's box, install black beam tube
* FI camera mount - make sure it's here, baked
* active TT pitch damping plan
* passive TT weight addition and pitch alignment plan
* camera plan for taking in-chamber photos without touching the tables
* look at layout of ITMY table. POY pick-off too close to main beam. Can we move POY pick-off to BS table?
* remount black glass baffles on SOS cages.
Stuff to bake:
* Quadrupus cables
* TT base plates
* FI mirror mount for FI camera
* stuff to add to active TTs
* stuff to add to passive TTs
|
7249
|
Wed Aug 22 15:47:34 2012 |
jamie | Summary | General | vent prepartion for fast-track vent |
We are discussing venting first thing next week, with the goal of
diagnosing what's going on in the PRC.
Reminder of the overall vent plan:
https://wiki-40m.ligo.caltech.edu/vent
Since we won't be prepared for tip-tilt installation (item 2), we should
focus most of the effort on diagnosing what's going on in the PRC. Of
the other planned activities:
(1) dichroic mirror replacement for PR3 and SR3
Given that we'll be working on the PRC, we might consider going ahead
with this replacement, especially if the folding mirror becomes
suspect for whatever reason. In any case we should have the new
mirrors ready to install, which means we should get the phase map
measurements asap.
(3) black glass beam dumps:
Install as time and manpower permits. We need to make sure all needed
components are baked and ready to install.
(4) OSEM mount screws:
Delay until next vent.
(5) new periscope plate:
Delay until next vent.
(6) cavity scattering measurement setup
Delay until next vent. |
7255
|
Thu Aug 23 10:38:12 2012 |
Steve | Update | General | vent prepartion for fast-track vent |
Quote: |
We are discussing venting first thing next week, with the goal of
diagnosing what's going on in the PRC.
Reminder of the overall vent plan:
https://wiki-40m.ligo.caltech.edu/vent
Since we won't be prepared for tip-tilt installation (item 2), we should
focus most of the effort on diagnosing what's going on in the PRC. Of
the other planned activities:
(1) dichroic mirror replacement for PR3 and SR3
Given that we'll be working on the PRC, we might consider going ahead
with this replacement, especially if the folding mirror becomes
suspect for whatever reason. In any case we should have the new
mirrors ready to install, which means we should get the phase map
measurements asap.
(3) black glass beam dumps:
Install as time and manpower permits. We need to make sure all needed
components are baked and ready to install.
(4) OSEM mount screws:
Delay until next vent.
(5) new periscope plate:
Delay until next vent.
(6) cavity scattering measurement setup
Delay until next vent.
|
Bob is back. Cleaning and baking all our posts and clamps. They will be ready for use Tuesday next week. Therefore beam dumps will be available for installation. |
5698
|
Wed Oct 19 14:03:41 2011 |
kiwamu | Update | General | vent prep : dichroic mirrors |
Status update on dichroic mirror:
I got the specification sheet of an aLIGO 2" dichroic mirror from Lisa.
https://dcc.ligo.org/cgi-bin/private/DocDB/ShowDocument?docid=25232
This is the one from ATF. It has low loss, a high R for 1064 nm and high T for 532 nm. So it matches our needs.
Based on this sheet we may reset some of the parameters in the specification (e.g. incident angle and etc.,) and will get a quote from ATF.
We will buy 3 of them, including 1 spare. First I need to review the specification. |
2636
|
Tue Feb 23 19:07:43 2010 |
kiwamu and steve | Configuration | VAC | vent is completed |
Quote: |
Kiwamu and Steve have started venting the 40m vacuum envelope.
Preparation:
centered oplevs at resonating cavities,
ITM references were set by green pointer from the ends by Koji,
closed PSL shutter and placed manual block into beam path,
checked jamnuts in locked positions on bellows,
turned HV off at PZT-Jena "steering mirror" power supply and OMC HV ps
checked particle counts,
switched oplev servos off,
set up N2 cylinder to start vent from 1e-6 Torr to 25 Torr,
have ~ 6 cylinders of instrument grade compressed air to bring envelope from 25 Torr to 760 Torr
All three cranes were wiped off today.
|
Kiwamu has completed the vent. |
10532
|
Wed Sep 24 13:28:48 2014 |
Steve | Update | VAC | vent is completed |
The vent is completed. ITMX was kicked up accidentally. Valve configuration: chamber open, RGA is pumped through VM2 Maglev |
10533
|
Wed Sep 24 16:02:58 2014 |
Jenne | Update | VAC | vent is completed |
[Steve, EricQ, Jenne]
ITMY and BS heavy doors are off, light doors are on. Q is aligning the IFO. |