  3796   Wed Oct 27 12:32:53 2010 josephbUpdateCDSfb rebooted to try and fix testpoints

Problem:

Test points were unavailable last night, even after reboots of c1sus and even restarting the daqd process on the frame builder.

Cause:

Its unclear at this time.  My guess is flaky fb and mx_stream codes.  At the moment, the daqd often requires several restarts as it segfaults within a minute or two of restarting it.

What we did (aka treating the symptoms):

We rebooted the frame builder machine.  I also added the daqd and nds processes to the inittab.  Now when these die, they will automatically be restarted.

Steps to add to the inittab on fb

0) If not on fb, ssh -X fb

1) cd /etc/

2) sudo vi inittab or sudo emacs init

3) Add a line like: id:runlevels:action:process

The id is a unqiue 2-4 letter and number identifier for the process

Run levels is the run level of linux that it will start at. 345 will cover the normal cases

action is what to do with the process. Respawn makes it run at startup and also restarts it everytime it dies.

process is the command you want to run

See "man inittab" for more details

In this case we added

daq:345:respawn:/opt/rtcds/caltech/c1/target/fb/daqd -c /opt/rtcds/caltech/c1/target/fb/daqdrc > /opt/rtcds/caltech/c1/target/fb/daqd.log


nds:345:respawn:/opt/rtcds/caltech/c1/target/fb/nds pipe > /opt/rtcds/caltech/c1/target/fb/nds.log

4) Save.

5) Run "sudo /sbin/telinit q".  This forces init to rexamine the inittab file

daqd and nds will now automatically restart when they die.

Continuing issues:

When the frame builder dies, the mx_stream processes on the front ends die as well.  These need to be restarted manually at the moment by using "sudo /etc/restart_streams" while on c1sus.

The framebuilder code shouldn't be this flaky.

  3797   Wed Oct 27 15:38:19 2010 josephb, yutaUpdateCDSIO chassis with bad timing was taken back to Downs

Problem:

The front end timing was not working properly for 2 of the IO chassis.  They were not being synced to the 1 PPS signal. 

This prevented the use of RFM for communication between front ends because time stamps on the transmitted data did not match the cycle on the receiving machine.

Action:

We took one of the incorrectly working chassis over to Downs.  Rolf said he would take a look at it tomorrow morning.

Joe will be going over tomorrow morning to talk with Rolf and see what needs to be done to fix it.

 

  3799   Wed Oct 27 17:06:48 2010 josephbUpdateCDSMoved c1iscey chassis and host interface board to c1ioo

Problem:

Need a working IO chassis connected to c1ioo in order to bring the MC_L into the digital realm, and then via RFM transmit to the c1sus machine.

Attempted Solution:

Move the c1iscey IO chassis to c1ioo while the c1ioo chassis is at downs.

The c1iscey chassis however doesn't seem to be talking to the c1ioo computer.  I tried changing the host interface card on the c1ioo chassis.  I took out One Stop Systems HIB2-x4-H interface card with serial number 26638 from the c1ioo computer and put in the One Stop Systems HIB2-x4-H with serial number 35242 in from c1iscey into c1ioo.  Still didn't work.

All the lights are red on the interface card on the actual chassis and its cooling fan isn't spinning. 

Using dmesg on c1ioo shows that it does not see any of the ADC/DAC/BO cards.

Status:

I'm going to  wait until tomorrow morning when Rolf gets a chance to look at the c1ioo chassis over at Downs to determine the next step.  If we fix the c1ioo chassis, I'm move the c1iscey chassis and its host interface board back to the end.

  3811   Thu Oct 28 16:38:54 2010 josephbUpdateCDSFlaky fb, reverted inittab changes on fb

Problem:

Yuta reported many of the signals being displayed by dataviewer "fuzzier" than normal.  And diaggui was not working.

Running "diag -i" reported:

Diagnostics configuration:
awg 21 0 192.168.113.85 822095893 1 192.168.113.85
awg 36 0 192.168.113.85 822095908 1 192.168.113.85
awg 37 0 192.168.113.85 822095909 1 192.168.113.85
tp 21 0 192.168.113.85 822091797 1 192.168.113.85
tp 36 0 192.168.113.85 822091812 1 192.168.113.85
tp 37 0 192.168.113.85 822091813 1 192.168.113.85

This seems to be missing an nds type line between the 3 awgs and the 3 tp lines.

The daqd code (the framebuilder) is being especially flaky today, and I'm starting to see new errors.

[Thu Oct 28 16:13:46 2010] Couldn't open full trend frame file
`/frames/trend/second/9723/C-

T-972342780-60.gwf' for writing; errno 13
epicsThreadOnceOsd epicsMutexLock failed.
epicsThreadOnceOsd epicsMutexLock failed.
epicsThreadOnceOsd epicsMutexLock failed.
epicsThreadOnceOsd epicsMutexLock failed.
epicsThreadOnceOsd epicsMutexLock failed.
Segmentation fault (core dumped)

or

[Thu Oct 28 16:17:06 2010] Couldn't open full frame file
`/frames/full/9723/.C-R-972343024-16.gwf' for writing; errno 13
CA client library tcp receive thread terminating due to a C++ exception
FATAL: exception not rethrown
CA client library tcp receive thread terminating due to a C++ exception
CA client library tcp receive thread terminating due to a C++ exception
FATAL: exception not rethrown
cac: tcp send thread received an unexpected exception - disconnecting
Aborted (core dumped)

What was done today that might have affected it:

A new c1ioo chassis from Downs was connected to c1ioo.  I also connected c1ioo to the DAQ network (192.168.114.xxx) which talks to the frame builder.

I started downloading the necessary files to be able to follow Keith's instructions for a standard control / teststand setup in /opt/apps , /opt/rtapps, etc.  However, it has not actually been installed yet.

Yuta added additional OL channels to the DAQ config for being recorded.

Attempted Fixes:

I reverted the inittab changes I made in this elog.  Didn't help.

I disconnected c1ioo from the DAQ network.  Didn't help.

Rebooted the frame builder machine.  Didn't help.

I've sent an e-mail to Alex describing the problem to see if he has any idea where we went wrong.

Yuta may try restoring the old DAQ channel choices and see if that makes a difference.

Current Status:

daqd framebuilder code still won't stay up.  So no channels at the moment.

  3814   Thu Oct 28 21:20:11 2010 yutaUpdateCDSFlaky fb, tried DAQ re-install, but no help

Summary:
  Unfortunately, fb is flakier than normal. We can't use dataviewer and diaggui now.
  I thought it might be because editting .ini files(list of DAQ channels) in /cvs/cds/rtcds/caltech/c1/chans/daq/ without using GUI was doing something wrong.
  So, I re-installed DAQ, but it didn't help.

What I did:
1. ssh c1sus, went to /opt/rtcds/caltech/c1/core/advLigoRTS/ and ran
  make uninstall-daq-c1SYS
  make install-daq-c1SYS

It didn't help.
More than that, MC suspension damping went wrong. So;

2. Rebooted c1sus machine.
 I restored MC suspension damping by doing this.
 (Similar thing happened Tuesday when we were trying to lock MC)

Conclusion:
  Editting .ini DAQ channel list file wasn't wrong. (or, I failed in finding anything wrong right now)

Quote:

Attempted Fixes:

Yuta may try restoring the old DAQ channel choices and see if that makes a difference.

 

  3815   Thu Oct 28 23:17:15 2010 yutaSummaryCDS[EMERGENCY] accidentally deleted daqd

Rana showed me that if c1sus machine runs c1mcs stuff(and c1x02 stuff) only, we can use dataviewer without crashing fb.
Also, if we set correct NDS server and port(fb/8088), we can use diaggui on every machine.


During my investigation on what he did, I accidentally deleted daqd......
I am very very sorry.

I don't know if it helps or not, but all I have is the following information:

[Backup?]
    /opt/rtcds/caltech/c1/target/fb/daqd.25sep10


[What I deleted]
   -rwxr-xr-x 1 controls controls 6583071 Oct  1 11:57 daqd


[help message for daqd existed]
CDS Data Acquisition Server, Frame Builder, version 2.0
California Institute of Technology, LIGO Project
Client communication protocol version 11.4

Usage:
        daqd [-f <input frame file name>]
        [-c <configuration file (default -- $HOME/.daqdrc)>]

        [-s <frame writer pause usec (default -- 1 sec)>]


This executable compiled on:
        Fri Oct  1 10:33:18 PDT 2010
        Linux fb 2.6.34.1 #7 SMP Fri Sep 24 14:09:53 PDT 2010 x86_64 Dual-Core AMD Opteron(tm) Processor 8220 AuthenticAMD GNU/Linux



Please help me, Joe.

  3821   Fri Oct 29 11:25:15 2010 josephb, yutaSummaryCDS[EMERGENCY] accidentally deleted daqd

Problem:

Missing daqd file, i.e. the framebuilder executable.

Solution:

1) Go to /opt/rtcds/caltech/c1/core/advLigoRTS/

2) Look in the Makefile for a likely build suspect.  In this case it was build dc, which stands for data concentrator.

3) So we ran "make dc"

4) Go to the sub-directory build/dc/ and then copy the daqd file there to the /opt/rtcds/caltech/c1/target/fb directory

5) Test it to ensure we're getting channels (Yay!)

Future Safeguards:

Place the new target directory under SVN control.

 

  3822   Fri Oct 29 11:29:29 2010 josephbUpdateCDSHow I broke the frame builder yesterday

Problem:

Long before Yuta came along and deleted daqd, I had done something to prevent the framebuilder code from running at all.

Cause:

Alex pointed out via e-mail that the corresponded to the inability to access certain frame files due to their permissions being only for root. 

Turns out when I had run the code under the inittab, I forgot to make it use controls, instead of root (which is the default).  This later on caused problems for the code when it tried to access those files, resulting in the wierd errors we saw.

Solution:

Use chown to change the offending frame files back to controls.

Future:

Write a proper inittab script which uses "su controls" before running the daqd code.

  3827   Fri Oct 29 16:43:25 2010 josephbUpdateCDSc1ioo now talking to c1sus

Problem:

c1ioo was not able to talk to c1sus because of timing issues.  This prevented the mode cleaner length signal (MC_L) from getting to c1sus.

Solution:

The replacement c1ioo chassis from Downs with a more recent revision of the IO backplane works. 

The c1ioo is now talking to c1sus and transmitting a signal.

We connected the cable hanging off the DAQ interface board labeled MC OUT1 to the MC Servo board's output labeled OUT1.

During debugging I modified the c1x02, c1x03, c1mcs and c1ioo codes to print debugging messages.  This was done by modifying the /opt/rtcds/caltech/c1/advLigoRTS/src/fe/commData2.c file.  I have since reverted those changes.

Future:

We still need to check that everything is connected properly and that the correct signal is being sent to the MC2 suspension.

 

  3828   Fri Oct 29 18:37:33 2010 yutaSummaryCDSdaqd and current CDS status

Background:
  Before Joe left(~ 1 hour ago), fb was working for a while. But after he left, daqd core dumped.
  This is maybe because we started c1sus and c1rms again for a delay measurement, just before he left.

What I did:
  I restarted IOP(c1x02) and FE models.
  Now it seems OK (we can use dataviewer and diaggui), but daqd reports bunch of errors like;

CA.Client.Exception...............................................
    Warning: "Identical process variable names on multiple servers"
    Context: "Channel: "C1:SUS-ITMX_TO_COIL_0_3_INMON", Connecting to: 192.168.113.85:42367, Ignored: c1susdaq:42367"
    Source File: ../cac.cpp line 1208
    Current Time: Fri Oct 29 2010 18:07:39.132686519
..................................................................


tp node xx invalid  (xx is 38 to 36)

Current CDS status:

MC damp dataviewer diaggui AWG c1ioo c1sus c1iscex RFM Sim.Plant  ...
                   ... 

   Please add other stuff you need.

Below is an example of how the color code works:

06-25-2009.NN_24ThreatLevel.GJH2L69BK.1.jpg

  3829   Sat Oct 30 05:27:53 2010 yutaSummaryCDSCDS time delay measurement

Motivation:
  We want to know the time delay of CDS in the IOP scheme.

Setup:
delaysetup.png

What I did:
1. Plugged out SCSI cable from ADC card 2 and DAC card 0 on C1SUS machine.
   ADC card 2 is ADC 0
   DAC card 0 is DAC 0

2. Measured tranfer function between ADC and DAC by SR785 and compared with the downsampling filter in IOP with 65534Hz(=4x16384Hz) sampling frequency.

  As ADC_0_0 corresponds to PRM ULSEN input and DAC_0_0 corrsponds to ULCOIL output, we turned all the filters off and set gains to 0 or 1 so that TF between ULSEN to ULCOIL will be ideally 1. (see this wiki page for channel assigns)

  The filter coefficients for the down sampling filter was found in;
    /cvs/cds/rtcds/caltech/c1/core/advLigoRTS/src/fe/controller.c
  It was named feCoeff4x.

static double feCoeff4x[9] =
        {0.014805052402446,
        -1.71662585474518,    0.78495484219691,   -1.41346289716898,   0.99893884152400,
        -1.68385964238855,    0.93734519457266,    0.00000127375260,   0.99819981588176};


3. Calculated the time delay dt using the following formula;
  dt = [pm - pc]/f/360deg    (pm: measured phase, pc: calculated phase from feCoeff4x, f: frequency)

4. Measured TF between the SCSI cables to estimate the effect of the cables and others.
  Disconnected SCSI cables from ADC and DAC, and connected A aad B(see setup).
  I measured both when input coupling of SR785 is DC and AC and see what happens.

Result:
  [time delay of the CDS]  (left, middle)
    The time delay gets larger with frequency. The time delay seems to be -175 usec at DC.
    However, the gain seems a little different from my expectation(feCoeff4x). So, there are maybe other filters I don't know.
    I neglected TF of upsampling this time.

  [cable and other effect]  (right)
    The effect to the time delay measurement was tiny by a factor of 10^4 to 10^3 (few nsec).
    But the total cable length was about 5 m and assuming signal speed is 0.6c, delay will be about 30 nsec.
    I don't know what's happening.

CDSdelay.png

Plan:
  - make a model that does not go through IOP and see the delay caused by IOP

By the way:
  fb daqd is still running for hours!
  Every FEs are running(c1sus,rms,mcs).

  3830   Sat Oct 30 14:35:43 2010 KojiSummaryCDSCDS time delay measurement

Unsatisfactory.

Neglecting the digital anti-imaging filter makes the discrepancy. You must take into account your digital filter twice.

I attached the slides I made during my visit for March LVC '09. P.5 would be useful.

Quote:

Result:
  [time delay of the CDS]  (left, middle)
    The time delay gets larger with frequency. The time delay seems to be -175 usec at DC.
    However, the gain seems a little different from my expectation(feCoeff4x). So, there are maybe other filters I don't know.
    I neglected TF of upsampling this time.

 

Attachment 1: CDS_system_investigation_090323.pdf
CDS_system_investigation_090323.pdf CDS_system_investigation_090323.pdf CDS_system_investigation_090323.pdf CDS_system_investigation_090323.pdf CDS_system_investigation_090323.pdf CDS_system_investigation_090323.pdf CDS_system_investigation_090323.pdf CDS_system_investigation_090323.pdf
  3834   Mon Nov 1 11:34:08 2010 josephbUpdateCDSCA.Client.Exception spam fixed

Problem:

CA.Client.Exception...............................................
    Warning: "Identical process variable names on multiple servers"
    Context: "Channel: "C1:SUS-ITMX_TO_COIL_0_3_INMON", Connecting to: 192.168.113.85:42367, Ignored: c1susdaq:42367"
    Source File: ../cac.cpp line 1208
    Current Time: Fri Oct 29 2010 18:07:39.132686519

The above exception gets thrown for each channel sent to the framebuilder when you start the frame builder.  Its the 192.168.113.xxx (main martian) and 192.168.114.xxx (DAQ) networks sending the same information.  This makes it hard to see real errors when starting the frame builder.

Solution:

Configure EPICS channel access to only use one network.  This is done by modifying the /etc/bash/bashrc and the /diskless/root/etc/bash/bashrc files with the following two lines:

export EPICS_CA_AUTO_ADDR_LIST=NO
export EPICS_CA_ADDR_LIST="192.168.113.255"

The first line tells the computers not to automatically search all attached network devices.  The second tells it to use the 192.168.113.xxx network for EPICS channel broadcasts.

  3836   Mon Nov 1 14:53:30 2010 josephb, alex, rolfUpdateCDSAlex updated FB and mx_streams code

Problem:

The framebuilder was being flaky.  MX_streams would go down, prevent testpoints from working and so forth.

Solution:

Send Alex up North for a week to fix the code. 

Alex came back and installed updates to the frame builder and the mx_streams code (Myrinet Express over Generic Ethernet Hardware) used by the front ends to talk to the frame builder.  Instead of 1 stream per model, there's now just 1 per front end handling all communications.

Alex did an SVN update and we now have the latest CDS code.

Self restarting codes:

The frame builder code (daqd) and nds pipe have been added to the fb machine's inittab.  Specifically it calls a script called /opt/rtcds/caltech/c1/target/fb/start_daqd.inittab and /opt/rtcds/caltech/c1/target/fb/start_nds.inittab.

The addition to the /etc/inittab file on fb is:

daq:345:respawn:/opt/rtcds/caltech/c1/target/fb/start_daqd.inittab 

nds:345:respawn:/opt/rtcds/caltech/c1/target/fb/start_nds.inittab

When these codes die they should automatically restart.

Self starting codes at boot up:

The front ends now start the mx_stream script (which lives in /opt/rtcds/caltech/c1/target/fb/ directory) at boot up.  They call it with the approriate command line options for that front end. It can be found in the /etc/rc.local file.

They look like: mx_stream -s "c1x02 c1sus c1mcs c1rms" -d fb:0

As always, the front end codes to be started are defined in the /etc/rtsystab file (or on fb, in the /diskless/root/etc/rtsystab file).

However, if it does go down you would need to restart it manually, although it seems more robust now and doesn't seem to go down every time we restart the frame builder.

All the usual front end IOCs and modules should be started and loaded on boot up as well.

 

Current CDS status:

MC damp dataviewer diaggui AWG c1ioo c1sus c1iscex RFM Sim.Plant Frame builder  ...
                     ... 
  3837   Mon Nov 1 15:35:41 2010 josephbUpdateCDSFront end USR and CPU times now recorded by DAQ

Problem:

We have no record of how long the CPUs are taking to perform a cycle's worth of computation

Solution:

I added the following channels to the various slow DAQ configuration files in /opt/rtcds/caltech/c1/chans/daq/

IOO_SLOW.ini:[C1:FEC-34_USR_TIME]
IOO_SLOW.ini:[C1:FEC-34_CPU_METER]
IOP_SLOW.ini:[C1:FEC-20_USR_TIME]
IOP_SLOW.ini:[C1:FEC-20_CPU_METER]
IOP_SLOW.ini:[C1:FEC-33_USR_TIME]
IOP_SLOW.ini:[C1:FEC-33_CPU_METER]
MCS_SLOW.ini:[C1:FEC-36_USR_TIME]
MCS_SLOW.ini:[C1:FEC-36_CPU_METER]
RMS_SLOW.ini:[C1:FEC-37_USR_TIME]
RMS_SLOW.ini:[C1:FEC-37_CPU_METER]
SUS_SLOW.ini:[C1:FEC-21_USR_TIME]
SUS_SLOW.ini:[C1:FEC-21_CPU_METER]

 

Notes:

To restart the daqd code, simply kill the running process.  It should restart automatically.  If it appears not to have started, check the /opt/rtcds/caltech/c1/target/fb/restart.log file and the /opt/rtcds/caltech/c1/target/fb/logs/daqd.log.xxxx files.  If you made a mistake in the DAQ channels and its complaining, fix the error and then restart init on the fb machine by running "sudo /sbin/init q"

  3838   Mon Nov 1 15:47:15 2010 yutaSummaryCDSCDS time delay measurement

Background:
  I measured CDS time delay last week, but because of my lack of understanding the system, it was incorrect.
  IOP has an anti-aliasing filter before downsampling from 64kHz(65536Hz) to 16kHz(16384Hz) and also has an anti-imaging filter before upsampling from 16kHz to 64kHz.
  So, I should have take feCoeff4x into account twice.
downupsampling.png

Result:
  TF agreed well with 2-time feCoeff4x and CDS time delay was -123.5 usec.
CDSdelay2.png


Plan:
 - make AWG(, diaggui TF measurement, tdssine) work
 - check input/output filter switching (using tdssine & tdsdmd)
 - measure openloop TF of MC suspension damping
 - divide it with my expectation and see if there are any filters I don't know

Quote:

Unsatisfactory.

Neglecting the digital anti-imaging filter makes the discrepancy. You must take into account your digital filter twice.

I attached the slides I made during my visit for March LVC '09. P.5 would be useful.

 

  3839   Mon Nov 1 16:43:24 2010 KojiSummaryCDSCDS time delay measurement

Um, Beautiful.

Actually, 123.5usec is almost exactly twice of 1/16384Hz.
Because of the loop, we have 1/16384Hz delay. I wonder where we do have the delay.

In order to understand the behaviour of the system can I ask you to test the following things?

1) What are the delay without IOPs with fsampl of 16k, 32k, 64k?

2) What are the delay with IOP with fsampl of 32k, 64k?

Quote:

Result:
  TF agreed well with 2-time feCoeff4x and CDS time delay was -123.5 usec.
CDSdelay2.png

 

  3841   Mon Nov 1 19:32:08 2010 yutaSummaryCDSfb crashed? during c1ioo and c1mcs connection at ASC

Frame builder died again!!

Background:
  We want to do angle to length measurement to optimize the beam position and increase visibility of MC locking.
  In order to do A2L measurement, we need excitation point, but AWG is currently not working.
  The better way is to use LOCKIN stuff like we had for OMC and put it to C1IOO WFS.
  A software oscillator in LOCKIN shakes the suspension, and demodulate the length signal.
  We can choose whatever DOF to shake, whatever signal to demodulate. It would be useful not just for A2L.

What I did:

  I started to put C1IOO WFS signal into C1SUS MC suspension RT model, but after compiling new c1mcs, fb crashed.
  Looks like daqd and mx_streams are running, but DAQ is not working(red).
  I don't know how to restart in a new way!

  3842   Mon Nov 1 23:31:05 2010 yutaUpdateCDSchecked input hardware filter in single frequency

Background:
  For input filter, we have analog whitening filter and also digital whitening filter. They have the same TF and when analog one is off, digital one should be on and vice versa.
  I made a python script that checks the switching automatically.

Method:
  Excite the suspension in a single frequency and see sensor inputs(XXSEN_IN1).
  Calculate the magnitude in the excitation frequency and compare it when digital whitening is off and on.
  When digital whitening is off, analog should be on, so sensor inputs should gone though the analog filter. That means the signal is multiplied by the TF of that filter, which makes the difference.

  We currently don't have excitation and I thought I have to wait, but instead of putting some extra excitation, I found that 60Hz line noise is useful.

Script:
  The script is /cvs/cds/caltech/users/yuta/scripts/WDWchecker.py
  For every sensor input, it;
    0. Stores current filter switching(XXSEN_SW1R)
    1. turns OFF the digital filter(FM1, using ezcaswitch)
    2. tdsdmd XXSEN_IN1 in 60Hz
    3. turns ON the digital filter
    4. tdsdmd XXSEN_IN1 in 60Hz
    5. divides mag(2.) by mag(4.) and calculate the analog filter gain in 60Hz
    6. Restores the filter switching in the state before the checking

Result:
  The results are;

C1:SUS-BS_ULSEN_IN1: 22.2 dB
C1:SUS-BS_URSEN_IN1: 18.7 dB
C1:SUS-BS_LRSEN_IN1: 22.7 dB
C1:SUS-BS_LLSEN_IN1: 16.0 dB
C1:SUS-BS_SDSEN_IN1: 21.5 dB
C1:SUS-ITMX_ULSEN_IN1: 16.9 dB
C1:SUS-ITMX_URSEN_IN1: 16.3 dB
C1:SUS-ITMX_LRSEN_IN1: 17.5 dB
C1:SUS-ITMX_LLSEN_IN1: 17.1 dB
C1:SUS-ITMX_SDSEN_IN1: 6.2 dB
C1:SUS-ITMY_ULSEN_IN1: 15.5 dB
C1:SUS-ITMY_URSEN_IN1: 16.5 dB
C1:SUS-ITMY_LRSEN_IN1: 17.4 dB
C1:SUS-ITMY_LLSEN_IN1: 16.3 dB
C1:SUS-ITMY_SDSEN_IN1: 18.0 dB
C1:SUS-PRM_ULSEN_IN1: 0.1 dB
C1:SUS-PRM_URSEN_IN1: 10.3 dB
C1:SUS-PRM_LRSEN_IN1: 13.1 dB
C1:SUS-PRM_LLSEN_IN1: -32.3 dB
C1:SUS-PRM_SDSEN_IN1: 14.6 dB
C1:SUS-SRM_ULSEN_IN1: 17.3 dB
C1:SUS-SRM_URSEN_IN1: 13.5 dB
C1:SUS-SRM_LRSEN_IN1: 1.6 dB
C1:SUS-SRM_LLSEN_IN1: 16.7 dB
C1:SUS-SRM_SDSEN_IN1: 18.3 dB

C1:SUS-MC1_ULSEN_IN1: 17.0 dB
C1:SUS-MC1_URSEN_IN1: 18.6 dB
C1:SUS-MC1_LRSEN_IN1: 14.9 dB
C1:SUS-MC1_LLSEN_IN1: 27.0 dB
C1:SUS-MC1_SDSEN_IN1: 16.6 dB
C1:SUS-MC2_ULSEN_IN1: 19.8 dB
C1:SUS-MC2_URSEN_IN1: 14.0 dB
C1:SUS-MC2_LRSEN_IN1: 20.8 dB
C1:SUS-MC2_LLSEN_IN1: 16.1 dB
C1:SUS-MC2_SDSEN_IN1: 17.3 dB
C1:SUS-MC3_ULSEN_IN1: 15.5 dB
C1:SUS-MC3_URSEN_IN1: 17.3 dB
C1:SUS-MC3_LRSEN_IN1: 18.2 dB
C1:SUS-MC3_LLSEN_IN1: 18.7 dB
C1:SUS-MC3_SDSEN_IN1: 16.8 dB


  Whitening filter has 18dB gain at 60Hz. (It's 3Hz pole, 30Hz zero, 100Hz zero and 0dB at DC)
  So, from the result, at least MC suspensions look like they have correct switching.
  But some channels doesn't look ok.
  We have to check those.

Plan:
   - check ITMX_SDSEN, PRM_ULSEN, PRM_LLSEN, SRM_LRSEN input filters
   - check the script and see if the script can really check. maybe the script needs some adjustments (# of averaging, multiple frequency, ......)
   - make AWG(, tdssine) work
   - check output hardware filter

By the way:
  fb is back. I don't know why. With help from Joe, I just compiled c1mcs again and again changing number of RFM channels.

  3844   Tue Nov 2 11:34:53 2010 josephb, alexUpdateCDSdiagconfd running, excitations back in dtt

Problem:

Diagnostic test tools was starting with errors.

Cause:

After the reboot of the frame builder machine yesterday by Alex, the diagconfd daemon was not getting started by xinetd.  There was a sequence error in the startup where xinetd was being called before mounting drives from linux1.

Important Note:

If you do not see the "nds" line you would not have diagnostic tests enabled in the DTT:

[controls@rosalba apps]$ diag -i | grep nds
nds * * 192.168.113.202 8088 * 192.168.113.202

Solution:

Alex changed /etc/xinetd.d/diagconfd file to point to /opt/apps/gds/bin/diagconfd instead of /opt/apps/bin/diagconf.  He also ensured xinetd started after mounting from linux1.

Alex's Suggestion:
My feeling is we should get rid of this feature and have an NDS address
entry box in the "Online" tab in the DTT with the default "nds". I
mentioned this to Jim Batch and he greed with me, so maybe he is going to
implement this. So maybe you guys want to request the same thing too, send
the request to Rolf and Jim, so we can have the last demon exercised.

  3845   Tue Nov 2 13:51:40 2010 josephb, yutaUpdateCDSRFM slowdown problem

Problem:

Each RFM memory location which needs to be read by a front end model slows the model significantly.

With no RFM memory locations to be read (replaced with grounds), the c1mcs model runs around 25 microseconds per cycle.

With 1 RFM memory location (MC_L), it runs around 29-33 microseconds.

With 3 RFM memory locations (MC_L, MC1_PIT, MC1_YAW), it runs around 45 microseconds.

With 7 RFM memory locations, the code generally doesn't run at all, going past the 62 microsecond maximum required to be able to keep up with the 16 kHz sample rate.

Last night Yuta somehow got it running with 7 RFM memory locations, but in that case, all the odd numbered RFM channels (1,3,5 as counted by the ipc file) did not work.  It was running at around 55 microseconds in that case.

The c1ioo code which is writing the data to the RFM card is experiencing no such slow down.

Current CDS status:

MC damp dataviewer diaggui AWG c1ioo c1sus c1iscex RFM Sim.Plant Frame builder  ...
                     ... 
  3846   Tue Nov 2 15:24:18 2010 josephbUpdateCDSc1ioo and c1mcs only sending MC_L, MC1_PIT, MC1_YAW

In order to have the c1mcs model run, we're running with only 3 RFM channels between c1ioo and c1mcs at the moment.  This leaves the model at around 45 microseconds, and at least lets us damp.

Alex and I still need to track down why the RFM read calls are taking so much time to execute.

  3849   Wed Nov 3 02:23:11 2010 yutaSummaryCDSchecking whitening filter board

Summary:
  Last night, I found that some of the input channels have wrong hardware filter switching(see elog #3842).
  So, to check the whitening board(D000210), I swapped the one with ok switching and bad switching.
  During the checking, I somehow broke the board.
  I fixed it, and now the status is the same as last night (or, at least look like the same).

What I did:
  1. Switching for SRM_LRSEN looked bad and every input channel for MC3 looked OK.
     So, I unplugged the whitening board for SRM (1X5-1-5B) and plugged it into MC3's place(1X5-1-8B).

  2. Ran WDWchecker.py for MC3. The switching seemed OK for every input channel, which means the whitening board was not the wrong one.

  3. Swapped back the whitening board as it was.

  4. Found MC3_ULSEN_OUT and MC3_LLSEN_OUT was keep showing negative value(they should be positive).

  5. Check the board and found that one of LT1125 for UL/LL was wrong (broken virtual ground).

  6. Replaced LT1125 and put the board back to 1X5-1-8B.

  7. Checked the board with WDWchecker.py and dataviewer 5-hour minute trend.
      The input signal came back to normal value(Attachment #1), MC3 damping working, input filter switching seems working

before LT1125 replacement after LT1125 replacement
C1:SUS-MC3_ULSEN_IN1: -2.4 dB [!]
C1:SUS-MC3_URSEN_IN1: 16.9 dB
C1:SUS-MC3_LRSEN_IN1: 15.4 dB
C1:SUS-MC3_LLSEN_IN1: -1.1 dB [!]
C1:SUS-MC3_SDSEN_IN1: 18.4 dB
C1:SUS-MC3_ULSEN_IN1: 18.2 dB
C1:SUS-MC3_URSEN_IN1: 17.6 dB
C1:SUS-MC3_LRSEN_IN1: 16.6 dB
C1:SUS-MC3_LLSEN_IN1: 17.1 dB
C1:SUS-MC3_SDSEN_IN1: 16.2 dB


Result:
  The whitening board seems OK.
  The wrong one is either wiring or RT model. Or, the checking script.

Attachment 1: MC3SEN.png
MC3SEN.png
  3852   Wed Nov 3 09:32:00 2010 ranaSummaryCDSEurocard board swapping

When swapping Eurocard boards, it is safest to first turn down the power supplies and then do the swapping.

Otherwise, it is sometimes the case that people plug in the board slowly and/or asymmetrically. This can cause the opamp to see one of the power supply rails without the other (e.g. +15 but not -15).

This can destroy some opamps. The true danger is that there may be damage to the board which you do not notice for several months, thereby leaving a timebomb for the next person.

Don't be an electronics terrorist!

  3853   Wed Nov 3 15:13:55 2010 josephbUpdateCDSTemporary RFM slow read work around

Problem:

Each RFM read in the c1mcs model is adding ~7 microseconds to the cycle time.  Adding too many pushes it over the 62 microsecond limit.

RFM writes do not have this problem.

Temporary Solution:

The fastest fix was to create a new front end model, called c1rfm, which does nothing but read in the MC1, MC2, MC3 PIT and YAW signals from the c1ioo machine, and then passes them along to the c1mcs model via shared memory, which is fast.

This means the data being sent is 2 cycles slow, one to go over the RFM, and one to go over shared memory.  It is running at 16384 cycles/second, so it shouldn't have much impact at the frequencies we use those channels for.

MC_L is still being sent directly to the c1mcs front end code via RFM.

Current Status:

The c1mcs model is running at  30-33 microseconds for CPU time.

The c1rfm model is running at 45-47 microseconds for CPU time.

All 7 channels, MC_L, MC1_PIT, MC1_YAW, MC2_PIT, MC2_YAW, MC3_PIT, MC3_YAW are responding.

The c1rfm model was added to the /diskless/root/etc/rtsystab file on the fb machine so that it automatically starts on the reboot of c1sus.

The USR and CPU time channels for c1rfm were added to the MCS_SLOW.ini file in /opt/rtcds/caltech/c1/chans/daq/ so that the framebuilder records them, namely:

[C1:FEC-38_USR_TIME]
[C1:FEC-38_CPU_METER]

The framebuilder was restarted to take these new channels into account.

Future:

Finish implementing and debugging the "round robin" RFM reader so as to not require a seperate model to be doing RFM reads in parallel.

Look into improving read speed by either merging timestamps and data into a single  or reading time stamp once every tenth or hundreth cycle, although this at best provides a factor of 2 improvement.

Check to see if RFM card being on the IO chassis or directly in the computer chassis makes a difference.

Get Alex and Rolf to improve RFM read speed.

  3855   Wed Nov 3 17:01:01 2010 josephbSummaryCDSComparison of RFM read times

Problem:

RFM reads are slow.  Rolf has said it should take 2-3 microseconds per read. 

c1sus is taking about 7 microseconds per read, twice as slow as Rolf's claim.

Hypothesis:

The RFM card is in the IO chassis, and is sharing the PCIe bus with 4 ADC cards, 3 DAC cards, 4 BO cards, and a BIO card.  Its possible this crowded bus is causing the reads to take even longer.

Test Results:

Compare read times between the c1sus computer, which has its RFM card in the IO chassis, to c1ioo, which has its RFM card in the computer.

c1ioo:

No RFM reads: 8 microseconds

3 RFM reads: 20 microseconds (~4 per read)

6 RFM reads: 32 microseconds (~4 per read)

c1sus:

No RFM read: 25 microseconds (bigger model)

1 RFM read: 33 microseconds (~8 per read)

3 RFM read: 45 microseconds (~7 per read)

6 RFM read: Over 62 microseconds, doesn't run.

Conclusion:

It looks like moving the RFM card may help by about a factor of 2 in read speed, although its still not quite what Alex and Rolf claim it should be.

The c1mcs and c1ioo models have been reverted to their normal operations.

 

  3857   Wed Nov 3 21:19:40 2010 yutaUpdateCDSput LOCKIN to c1ioo model and checked

(Joe, Yuta)

Summary:
  LOCKIN(consists of oscillator and demodulator) is needed for A2L measurement.
  So, we put LOCKIN to c1ioo model, whose outputs goes to c1mcs ASC.
  After that, I checked the functionality of LOCKIN by directly connecting DAC output for a coil to ADC input for MCL with BNC cable.

What we did:
[Putting LOCKING to c1ioo model]
1. Copied Simulink LOCKIN stuff(cdsOsc, Product, cdsPhase ...) from /cvs/cds/caltech/cds/rward-advLigo/src/epics/simLink/omc.mdl and put it into c1ioo model.

2. Copied MEDM screen file /cvs/cds/caltech/medm/c1/omc/C1OMC_LSC_LOCKIN.adl and modified it for our use.

[Checking LOCKIN]
3. Disconnected MC2_ULCOIL input to SOS Coil driver at 1X4-1-6A and checked the signal from software oscillator at c1ioo is coming.

4. Disconnected the cable labeled "MC OUT1" at 1X2 (which is MCL signal to ADC) and put MC2_ULCOIL output directly using long BNC cable.

5. Checked the functionality of LOCKIN by StripTool.
   The cable wiring did not conflict with my expectation.
   Software mixer is working.(frequency is doubled. X_SIN has offset and X_COS doesn't)

6. Put the cables back.

Result:

  Thanks to c1rfm, c1ioo and c1sus are talking without ADC timeout.
  Also, LOCKIN is working fine.

Attachment 1: Screenshot_C1IOO_LOCKIN.png
Screenshot_C1IOO_LOCKIN.png
  3860   Thu Nov 4 15:15:43 2010 josephbUpdateCDSModified feCodeGen.pl, fmseq.pl, and suspension screens

Feature Requested:

Have the CPU_meter change change color at various alarm levels.  These alarm levels have been set at 2/3 maximum for Minor alarm (yellow) and 9/10 maximum for Major alarm (red).

Implementation:

Rather than hand code each EPICS .db file to add the alarm files each time we rebuild the front ends, I decided to modify it at the source (since it strikes me as a generally useful alarm level for all front end codes).

First, I modified the feCodeGen.pl script.

I changed

print EPICS "OUTVARIABLE FEC\_$dcuId\_CPU_METER epicsOutput.cpuMeter int ai 0 field(HOPR,\"$rate\") field(LOPR,\"0\")\n";

to

print EPICS "OUTVARIABLE FEC\_$dcuId\_CPU_METER epicsOutput.cpuMeter int ai 0 field(HOPR,\"$rate\") field(LOPR,\"0\") field(HIGH,\"$two_thirds_rate\") field(HSV,\"MINOR\") field(HIHI,\"$     nine_tenths_rate\") field(HHSV,\"MAJOR\")\n";

I added the following two lines just before it as well:

$two_thirds_rate = int($rate * 2 / 3);

$nine_tenths_rate = int($rate * 9 / 10);

However, only the first four fields were actually added to the database file.  Apparently fmseq.pl, which populated the database, was hard coded to only handle up to 4 fields.

I modified the fmseq.pl script in /opt/rtcds/caltech/c1/core/advLigoRTS/src/epics/util/ so as to be able to handle up to 6 field values when writing EPICS .db files.

This change was accomplished by simply changing the following line

($junk, $v_name, $v_var, $v_type, $ve_type, $v_init, $v_efield1, $v_efield2, $v_efield3, $v_efield4 ) = split(/\s+/, $_);

to 

($junk, $v_name, $v_var, $v_type, $ve_type, $v_init, $v_efield1, $v_efield2, $v_efield3, $v_efield4, $v_efield5, $v_efield6 ) = split(/\s+/, $_);

everywhere it occurred.  There were something like 10 instances of it. Also, I added the two lines

        $vardb .= "    $v_efield5\n";
        $vardb .= "    $v_efield6\n";

after each set of

         $vardb .= "    $v_efield1\n";
         $vardb .= "    $v_efield2\n";
         $vardb .= "    $v_efield3\n";
         $vardb .= "    $v_efield4\n";

Lastly, I modified the CPU_METER bar graph on the C1SUS_DEFAULTNAME.adl screen (located in /opt/rtcds/caltech/c1/medm/master/) to use alarm levels, and then ran generate_master_screens.py.

 

  3861   Thu Nov 4 15:27:33 2010 josephb, yutaUpdateCDSc1ioo test points not working

Problem:

We can't access any of the c1ioo computer testpoints.  Dataviewer and DTT both fail to read any data from them.

According to diag -l (when run on Rosalba in /opt/apps), the testpoints are not being set.  Also at some point during the day when debugging, we also somehow messed up all the front end connections to the framebuilder.

Errors reported by dataviewer:

Server error 13: no data found
datasrv: DataWriteRealtime failed: daq_send: Illegal seek
Server error 12: no such net-writer
Server error 13: no data found
datasrv: DataWriteRealtime failed: daq_send: Illegal seek
Server error 12: no such net-writer
read(); errno=0
Server error 6532: invalid channel name
Server error 16080: unknown error
datasrv: DataWriteRealtime failed: daq_send: Illegal seek

Error reported by daqd.log:

[Thu Nov  4 13:29:35 2010] About to request `C1:IOO-MC1_PIT_IN1' 10022
on node 34
[Thu Nov  4 13:29:35 2010] Requesting 1 testpoints; tp[0]=10022; tp[1]=32531

[Thu Nov  4 13:29:35 2010] About to request `C1:SUS-MC2_SUSPOS_IN1'
10094 on node 36
[Thu Nov  4 13:29:35 2010] Requesting 1 testpoints; tp[0]=10094; tp[1]=32531

[Thu Nov  4 13:29:38 2010] ETIMEDOUT: test point `C1:IOO-MC1_PIT_IN1'
(tp_num=10022) was not set by the test point manager; request failed
[Thu Nov  4 13:29:38 2010] About to clear `C1:IOO-MC1_PIT_IN1' 10022 on node 34

Attempted Fixes:

Remove all daq related files: /opt/rtcds/caltech/c1/target/gds/param/tpchn_c1ioo.par and /opt/rtcds/caltech/c1/chans/daq/C1IOO.ini.  Rebuilt the front end code.

Double checked ethernet connection between c1ioo and the daq router and fb. 

Confirmed open mx was running on c1ioo. Confirmed awgtpman was running (although at one point I did find duplicate awgtpman running for c1x03, the IOP associated with c1ioo).

Rebooted the c1ioo machine. Confirmed all necessary codes came back.

Restarted the daqd process several times on the fb machine.

Current Status:

Framebuilder is running, and c1sus testpoints are available.  c1ioo test points are not.  Waiting to hear back from Alex on possible ideas.

  3863   Thu Nov 4 17:53:29 2010 yutaUpdateCDSprimitive python script for A2L measurement

Summary:
  I wrote a python script for A2L measurement.
 Currently it is really primitive, but I tested the basic functionality of the script.

 We already have A2L script(at /cvs/cds/rtcds/caltech/c1/scripts/A2L) that uses ezlockin, but python is more stable and easy to read.

A2L measurement method:
  1. Dither a optic using software oscillator in LOCKIN and demodulate the length signal by that frequency.
  2. Change coil output gains to change the pivot of the dithering and do step 1.
  3. Coil output gain set that gives the smallest demodulated magnitude tells you where the current beam spot is.

  Say you are dithering the optic in PIT and changing the coil gains keeping UL=UR and LL=LL.
  If the coil gain set UL=UR=1.01, LL=LR=-0.99 gives you demodulated magnitude 0, that means the current beam spot is 1% upper than the center, compared to 1/2 of UL-LL length.
  You do the same thing for YAW to find horizontal position of the beam.

Description of the script:
  Currently, the script lives at /cvs/cds/caltech/users/yuta/scripts/A2L.py
  If you run;
     ./A2L.py MC1 PIT
  it gives you vertical position of the beam at MC1.

  It changes the TO_COIL matrix gain by "DELTAGAINS", turns on the oscillator, and get X_SIN, X_COS from C1IOO_LOCKIN.
  Plots DELTAGAINS vs X_SIN/X_COS and fit them by y=a+bx+cx^2.(Ideally, c=0)
  Rotates (X_SIN, X_COS) vectors to get I-phase and Q-phase.
    (I,Q)=R*(X_SIN,X_COS)
  Rotation angle is given by;
    rot=arctan(b(X_COS)/b(X_SIN))
  which gives Q 0 slope(Ideally, Q=0).
  x-intercept of DELTAGAINS vs I plot gives the beam position.

Checking the script:
  1. I used the same setup when I checked LOCKIN(see elog #3857). C1:SUS-MC2_ULCOIL output goes directly to C1:IOO-LOCKIN_SIG input.

  2. Set oscillator frequency to 18.13Hz, put 18.13Hz band-pass filter to C1:IOO-LOCKIN_SIG filter module, and put 1Hz low-pass filter to C1:IOO-LOCKIN_X_SIN/X_COS filter modules.
        Drive frequency 18.13Hz is same as the previous script(/cvs/cds/rtcds/caltech/c1/scripts/A2L/A2L_MC2).

  3. Ran the script. Checked that Q~0 and rot=-35deg.

  4. Put phase shifting filter to C1:IOO-LOCKIN_SIG filter module and checked Q~0 and rotation angle.
     fitler rot(deg)
     w/o    -35
     +90deg  45
     -90deg  56
     -45deg -80

  5. Put some noise in C1:SUS-MC2_ULCOIL by adding SUSPOS feedback signal and ran the script.(Attachment #1)
      During the measurement, the damping servo was off, so SUSPOS feedback signal can be treated as noise.

Conclusion:
  The result from the test measurement seems reasonable.
  I think I can apply it to the real measurement, if MCL signal is not so noisy.[status: yellow]

Plan:
  - add calculating coherence procedure, averaging procedure to the script
  - add setting checking procedure to the script
  - apply it to real A2L measurement

Bay the way:
  Computers in the control room is being so slow (rossa, allegra, op440m, rosalba). I don't know why.

Attachment 1: a2ltest.png
a2ltest.png
  3869   Fri Nov 5 15:20:18 2010 josephb, alexSummaryCDS40m computer slow down solved

Problem:

The 40m computers were responding sluggishly yesterday, to the point of being unusable.

Cause:

The mx_stream code running on c1iscex (the X end suspension control computer) went crazy for some reason.  It was constantly writing to a log file in /cvs/cds/rtcds/caltech/c1/target/fb/192.168.113.80.log.  In the past 24 hours this file had grown to approximately 1 Tb in size. The computer had been turned back on yesterday after having reconnected its IO chassis, which had been moved around last week for testing purposes - specifically plugging the c1ioo IO chassis in to it to confirm it had timing problems.

Current Status:

The mx_stream code was killed on c1iscex and the 1 Tb file removed.

Computers are now more usable.

We still need to investigate exactly what caused the code to start writing to the log file non-stop.

Update Edit:

Alex believes this was due to a missing entry in the /diskless/root/etc/hosts file on the fb machine.  It didn't list the IP and hostname for the c1iscex machine.  I have now added it.  c1iscex had been added to the /etc/dhcp/dhcpd.conf file on fb, which is why it was able to boot at all in the first place.  With the addition of the automatic start up of mx_streams in the past week by Alex, the code started, but without the correct ip address in the hosts file, it was getting confused about where it was running and constantly writing errors.

Future:

When adding a new FE machine, add its IP address and its hostname to the /diskless/root/etc/hosts file on the fb machine.

  3870   Fri Nov 5 16:40:22 2010 josephb, alexUpdateCDSc1ioo now has working awgtpman

Problem:

We couldn't set testpoints on the c1ioo machine.

Cause:

Awgtpman was getting into some strange race condition.  Alex added an additional sleep statement and shifted boot order slightly in the rc.local file.  This apparently is only a problem on c1ioo, which is a Sun X4600.  It was using up 100% of a single CPU for the awgtpman process.

Status:

We now have c1ioo test points working.

Future:

Need to examine the startc1ioo script and see if needs a similar modification, as that was tried at several points but yielded a similar state of non-functioning test points. For the moment, reboot of c1ioo is probably the best choice after making modifications to the c1ioo or c1x03 models.

Current CDS status:

MC damp dataviewer diaggui AWG c1ioo c1sus c1iscex RFM Sim.Plant Frame builder  TDS
                     
  3872   Fri Nov 5 21:49:12 2010 ranaSummaryCDS40m computer slow down solved

Quote:

Problem:

The 40m computers were responding sluggishly yesterday, to the point of being unusable.

Cause:

The mx_stream code running on c1iscex (the X end suspension control computer) went crazy for some reason.  It was constantly writing to a log file in /cvs/cds/rtcds/caltech/c1/target/fb/192.168.113.80.log.  In the past 24 hours this file had grown to approximatel

The moral of the story is, PUT THINGS IN THE ELOG. This wild process is one of those things where people say 'this won't effect anything', but in fact it wastes several hours of time.

  3877   Sat Nov 6 16:13:14 2010 ranaSummaryCDS40m computer slow down solved

As part of the effort to debug what was happening with the slow computers, I disabled the auto MEDM snapshots process that Yoichi/Kakeru setup some long time ago:

https://nodus.ligo.caltech.edu:30889/medm/screenshot.html

We have to re-activate it now that the MEDM screen locations have been changed. To do this, we have to modify the crontab on nodus and also the scripts that the cron is calling. I would prefer to run this cron on some linux machine since nodus starts to crawl whenever we run ImageMagick stuff.

Also, we should remember to start moving the old target/ directories into the new area. All of the slow VME controls are still not in opt/rtcds/.

  3880   Mon Nov 8 14:30:15 2010 josephb, yutaUpdateCDSAdded LIGONDSIP setting to cshrc.40m

We added the following line to the cshrc.40m file in the 64-bit linux and 32-bit linux sections:

setenv LIGONDSIP fb

This allows codes like tdsdmd to work properly on the linux machines (seemed to already work fine on the solaris op440m without this change).

  3891   Thu Nov 11 04:32:53 2010 yutaSummaryCDSfound poor contact of DAC cable, previous A2L results were wrong

(Koji, Jenne, Yuta)

We found one of DAC cables had a poor contact.
That probably caused our too much "tilt" of the beam into MC.

Story:
  From the previous A2L measurement and MC aligning, we found that the beam is somehow vertically "tilted" so much.
  We started to check the table leveling and the beam height and they looked reasonable.
  So, we proceeded to check coil balancings using optical levers.
  During the setup of optical levers, I noticed that VMon for MC1_ULCOIL was always showing -0.004 even if I put excitation to coils.
  It was because one of the DAC cables(labeled CAB_1Y4_88) had a poor contact.
  If I push it really hard, it is ok. But maybe we'd better replace the cable.

What caused a poor connection?:
  I don't know.
  A month ago, we checked that they are connected, but things change.

How to prevent it:
  I made a python script that automatically checks if 4 coils are connected or not using C1:IOO_LOCKIN oscillator.
  It is /cvs/cds/caltech/users/yuta/scripts/coilchecker.py.
  It turns off all 3 coils except for the one looking at, and see the difference between oscillation is on and off.
  The difference can be seen by demodulating SUSPOS signal by oscillating frequency.
  If I intentionally unplug CAB_1Y4_88, the result output for MC1 will be;

==RESULTS== (GPS:973512733)
MC1_ULCOIL      0.923853382664 [!]
MC1_URCOIL      38.9361304794
MC1_LRCOIL      55.4927575177
MC1_LLCOIL      45.3428533919


Plan:
 - Make sure the cable connection and do A2L and MC alignment again
 - Even if the cables are ok, it is better to do coil balancings. See the next elog.

  3895   Thu Nov 11 11:51:30 2010 KojiSummaryCDSfound poor contact of DAC cable, previous A2L results were wrong

The cause is apparent! The connectors on the cables are wrong!
Currently only 50% of the pin length goes into the connector!

Quote:

(Koji, Jenne, Yuta)

We found one of DAC cables had a poor contact.
That probably caused our too much "tilt" of the beam into MC.

  It was because one of the DAC cables(labeled CAB_1Y4_88) had a poor contact.
  If I push it really hard, it is ok. But maybe we'd better replace the cable.

What caused a poor connection?:
  I don't know.
  A month ago, we checked that they are connected, but things change.

 

  3900   Thu Nov 11 21:07:49 2010 josephbUpdateCDSPlugged c1iscex into DAQ network - still causes network slowdown

I connected the c1iscex computer to the dedicated DAQ network switch (located in 1X7).

This does not seem to have helped c1iscex stop spewing out "OMX: Failed to find peer index of board 00:00:00:00:00:00 (Peer Not Found in the Table)" at the rate of ~1 Gigabyte per minute.

c1iscex is currently off until a solution can be found.

  3901   Thu Nov 11 23:35:23 2010 KojiSummaryCDSfound poor contact of DAC cable, previous A2L results were wrong

[Koji / Yuta]

There were the guys who used the PENTEK 40pin connectors into the IDC 40pin connectors.
Those connectors are not compatible at all.

==> We replaced the connectors on the cables from DAC to IDC adapters to the dewhitening board for the vertex SUSs.

In addition, I found one of the Binary OUT IDC50pin connector has no clamp.

==> We put the IDC50pin clamp on it.

bad-boys.png

PENTEK connectors were inserted. The latches are not working!

IMG_3698.jpg

the vertical pitch is different between PENTEK and IDC!
IMG_3705.jpg

Wow! Where is the clamp???

IMG_3706.jpg

Quote:

The cause is apparent! The connectors on the cables are wrong!
Currently only 50% of the pin length goes into the connector!

Quote:

(Koji, Jenne, Yuta)

We found one of DAC cables had a poor contact.
That probably caused our too much "tilt" of the beam into MC.

  It was because one of the DAC cables(labeled CAB_1Y4_88) had a poor contact.
  If I push it really hard, it is ok. But maybe we'd better replace the cable.

What caused a poor connection?:
  I don't know.
  A month ago, we checked that they are connected, but things change.

 

 

  3906   Fri Nov 12 10:49:34 2010 josephb, valeraUpdateCDSTest of ADC noise

Test:

Look at the effects of the ADC voltage range on the ADC noise floor.

ADC input was terminated with 50 ohms.  We then looked at the channel with DTT. This was at +/- 10 V range.  We used C1:SUS-PRM_SDSEN_IN1 as the test channel.

The map.c file (in /opt/rtcds/caltech/c1/core/advLigoRTS/src/fe/ ) then had two lines added at line 766.

//JCB temporary 2.5V test, remove me
  adcPtr[devNum]->BCR &= 0x84240;

This hard coded the 2.5 V range (we default to the 10 V range at the moment).

We then rebuilt the c1x02 model and reran the test.

Finally, we reverted the code change to map.c and rebuilt c1x02.

Results:
I've attached the DTT output of the two tests.

It appears the ADC is limited by 1.6 uV/rtHz.  Hence the increase in noise in counts by a factor of 4 when we drop to +/- 2.5 V from +/- 10 V.

Attachment 1: ADC_noise.pdf
ADC_noise.pdf
  3910   Fri Nov 12 19:24:56 2010 KojiUpdateCDSTest of ADC noise

[Koji Yuta]

We found one of the ADC cables were left unconnected. This left the MC suspensions uncontrollable through the whole afternoon.
Please keep the status updated and don't forget to revert the configuration...

Quote:

Test:

Look at the effects of the ADC voltage range on the ADC noise floor.

ADC input was terminated with 50 ohms.  We then looked at the channel with DTT. This was at +/- 10 V range.  We used C1:SUS-PRM_SDSEN_IN1 as the test channel.

The map.c file (in /opt/rtcds/caltech/c1/core/advLigoRTS/src/fe/ ) then had two lines added at line 766.

//JCB temporary 2.5V test, remove me
  adcPtr[devNum]->BCR &= 0x84240;

This hard coded the 2.5 V range (we default to the 10 V range at the moment).

We then rebuilt the c1x02 model and reran the test.

Finally, we reverted the code change to map.c and rebuilt c1x02.

Results:
I've attached the DTT output of the two tests.

It appears the ADC is limited by 1.6 uV/rtHz.  Hence the increase in noise in counts by a factor of 4 when we drop to +/- 2.5 V from +/- 10 V.

 

  3912   Sat Nov 13 15:53:05 2010 yutaUpdateCDSdiagonalization of MC input matrix

Motivation:
  MC is aligned from the A2L measurement, but to do the beam centering more precisely, we need coils to be balanced.
  There are several ways to balance the coils, like using oplev or WFS QPD RF channels.
  But oplev takes time to setup, especially for MC3. Also, c1ioo WFS channels were newly setup and haven't been checked yet.
  So, I decided to use OSEM sensors.
  An OSEM sensor itself is sensitive to every DOF of an optic motion, but we can diagonalize them using 4 OSEM sensors and proper input matrix.

Method:

  1. Measure transfer functions between
     ULSEN and URSEN (H_UR(f))
     ULSEN and LRSEN (H_LR(f))
     ULSEN and LLSEN (H_LL(f))

  2. Make a matrix A.

    A =  [[ 1           1           1          ]
          [ H_UR(f_pos) H_UR(f_pit) H_UR(f_yaw)]
          [ H_LR(f_pos) H_LR(f_pit) H_LR(f_yaw)]
          [ H_LL(f_pos) H_LL(f_pit) H_LL(f_yaw)]]


    where f_dof are resonant frequencies.

  3. A is

    s = Ad


   where vectors s^T=[ULSEN URSEN LRSEN LLSEN] and d^T=[POS PIT YAW].
   So,

    d = Bs = (A^TA)^(-1)A^Ts

   where A^T is transpose of A.

   B is the input matrix that diagonalizes 3 DOFs.

What I did:

  1. Measured the TFs using diaggui and exported as ASCII.

  2. Made a script that reads that TF file, calculates and sets a new input matrix B.
    /cvs/cds/caltech/users/yuta/scripts/inputmatrixoptimizer.py
   You need to set resonant frequencies to use the script.

   New input matrices for MCs are;

C1:SUS-MC1_INMATRIX
[[ 1.17649712  0.94315611  0.85065054  1.02969624]
 [ 0.55939288  1.28066594 -0.85235358 -1.3075876 ]
 [ 1.23467139 -0.74521928 -1.29394051  0.72616882]]

C1:SUS-MC2_INMATRIX
[[ 1.12630748  1.01451545  0.9013457   0.95783137]
 [ 1.03043025  0.67826036 -1.37270598 -0.91860341]
 [ 0.83546271 -1.26311029 -0.6456881   1.2557389 ]]

C1:SUS-MC3_INMATRIX
[[ 1.18212117  1.26419447  0.77744155  0.77624281]
 [ 0.79344415  0.84959646 -1.10946339 -1.247496  ]
 [ 1.00225331 -0.84807863 -1.21772132  0.93194674]]


  I ignored SIDE this time.

Result:

  Spectra of each SUSDOF_IN1_DAQ before diagonalization (INMATRIX elements all 1 or -1) were
MCspectraNov09.png

  After diagonalization, spectra are
MCspectraNov13.png

  As you can see, each SUSDOF has only single peak (and SIDE peak) after the diagonalization.
  SUSSIDE still has 4 peaks because SIDE is not included this time.

  For MC2, POS to SUSPIT and POS to SUSYAW got worse. I have to look into them.

Effect of resonant frequency drift:

  As you can compare and see from the spectra above, resonant frequencies of MC1 are somehow drifted(~0.5%) from Nov 9 to Nov 13.
  If resonant frequency you expected was wrong, calculated input matrix will be also wrong.
  The effect of 0.5% drift and wrong input matrix can be seen from this spectra. DOFs are not clearly separated.
 MC1spetra_wrongmatrix.png

Plan:

 - learn how to use diaggui from command line and fully automate this process
 - balance the coils using these diagonalized SUSPOS, SUSPIT, SUSYAW

  3915   Sun Nov 14 11:56:59 2010 valeraUpdateCDSTest of ADC noise

 

We missed a factor of 2 in the ADC calibration: the differential 16 bit ADC with +/-10 V input has 20 V per 32768 counts (1 bit is for the sign). I confirmed this calibration by directly measuring ADC counts per V.

So the ADC input voltage noise with +/-10V range around 100 Hz is 6.5e-3 cts/rtHz x 20V/32768cts =  4.0 uV/rtHz. Bummer. 

The ADC quantization noise limit is 1/sqrt(12 fs/2)=1.6e-3 cts/rtHz. Where the ADC internal sampling frequency is fs=64 kHz. If this would be the limiting digitization noise source then the equivalent ADC input voltage noise would be 1 uV/rtHz with +/-10 V range.

  3919   Mon Nov 15 11:13:12 2010 josephbUpdateCDSModified rc.local to not start mx_streams automatically

Problem:

c1iscex floods the network with about 1 gigabyte of error messages in a few seconds, writing to a log file in /opt/rtcds/caltech/c1/target/fb/logs/

Temporary change:

I commented out the following line in the rc.local file on the fb machine in the /diskless/root/etc/ directory:

#nice --20 ./mx_stream -s "$SYSTEMS" -d fb:0 >& logs/$HOSTNAME.log&

This disables the automatic start up of the mx_streams code on all the front ends.  This will prevent the network being brought to its knees by c1iscex while we debug the problem.

It also means on a reboot of the front ends, the mx_stream process needs to be started by hand until this change is reverted.

To do this, log into the front end and then change directory to /opt/rtcds/caltech/c1/target/fb

For c1sus, run:

./mx_stream -s c1x02 c1sus c1mcs c1rms c1rfm  -d fb:0

For c1ioo, run:

./mx_stream -s c1x03 c1ioo -d fb:0

 

  3926   Mon Nov 15 16:26:46 2010 josephbUpdateCDSc1iscex is now running and the network hasn't died

Problem:

c1iscex was spamming the network with error messages.

Solution:

Updated the front end codes to current standards (they were on the order of months out of date).  After fixing them up and rebuilding the codes on c1iscex, it no longer had problems connecting to the frame builder.\

Status:

I can look at test points for ETMX.  It is not currently damping however.

To Do:

Move filters for ETMX into the correct files. 

Need to add a Binary output blue and gold box to the end rack, and plug it into the binary output card.  Confirm the binary output logic is correct for the OSEM whitening, coil dewhitening, and QPD whitening boards. 

Get ETMX damped.

Figure out what we're going to do with the aux crate which is currently running y-end code at the new x-end.  Koji suggested simply swapping auxilliary crates - this may be the easiest.  Other option would be to change the IP address, so that when it PXE boots it grabs the x-end code instead of the y-end code.

Current CDS status:

MC damp dataviewer diaggui AWG c1ioo c1sus c1iscex RFM Sim.Plant Frame builder TDS
                     
  3935   Tue Nov 16 21:42:31 2010 ranaUpdateCDSScreen Time Fix
I learned today that the following python code will do a find/replace to fix the TIME string on any MEDM screen which has a whited out time field.
Previously, this field was sourced from the c1dscepics of c1losepics process. Now we have to get it from the IOO or SUS front ends

Here's the python code:

import re
o = open("output.adl","w")
data = open("test.adl").read()
o.write( re.sub("C0:TIM-PACIFIC_STRING","C1:FEC-34_TIME_STRING",data)  )
o.close()

Where 'output.adl' could be the same name as 'test.adl' if you want to
replace the existing file. Also FEC-34 just refers to which FE you're running.
It could, in principle, be any one of them.
 

The next step is to figure out how to apply this to all the files in a directory.

  3938   Wed Nov 17 10:39:20 2010 josephbUpdateCDSScreen Time Fix

An improved python code to apply a replacement to all *.adl files in a directory would be:

import re, os
files = os.listdir("./")
  for file in files:
    if ".adl" in str(file):
      data = open(file).read()
      o = open(file,"w")
      o.write( re.sub("C0:TIM-PACIFIC_STRING","C1:FEC-34_TIME_STRING",data)  )
      o.close()

Of course, this entire python script can be replaced with a single sed command:

sed -i 's/C0:TIM-PACIFIC_STRING/C1:FEC-34_TIME_STRING/g' *

A more complicated script could be written which looks for key identifiers either in the file header or inside the file to determine which front end is appropriate, using a dictionary like:

dcuid_dict = {"BS":21,"PRM":37,"SRM":37,"ITMX":21,"ITMY":21,"MC1":36,"MC2":36,"MC3":36,"ETMX":24,"ETMY":26}

and then using for loops and if statements.

 

  3940   Wed Nov 17 16:02:30 2010 josephbUpdateCDSModified feCodeGen.pl to fix filtMuxMatrix name generation

Problem:

Sometime in the last 3 weeks, probably when Alex brought his latest changes from Hanford to the 40m and did an SVN update, the code which generates the names of the filter .adl files links for the overall matrix view broke.

Fix:

I modified FE code gen to use $basename instead of the base name after the top name transform (this changes _ to - after the first 3 letters

@@ -3520,11 +3522,11 @@
 
                  my $tn = top_name_transform($basename);
                  my $basename1 = $usite . ":" . $tn . "_";
-                 my $filtername1 = $usite . $tn;
+                 my $filtername1 = $usite . $basename;

Still having problems:

The filter modules built with the matrix of filter modules run (offests/gains work), but will not load filter coefficients/filter names.  All the other filter modules outside the matrix seem to load fine.  At this point, doing a rebuild of any of the front end machines may cause the A2L filter banks to be unloadable.

 

  3941   Wed Nov 17 20:44:59 2010 yutaSummaryCDSno QPD channels on c1sus machine today

(Joe, Suresh, Yuta)

Currently, only 2 ADC cards work on c1sus machine.
No QPD inputs(e.g. MC2 trans), and no RFM.


Summary:
  We wanted to have PEM(physical environment montor) channels, so we moved a ADC card in c1sus machine.
  It ended up with destroying one of the 3 ADCs.

What we did:
  1. Moved ADC card at PCIe expansion board slot 0 to other empty slot.
     What we call PCI slot 0 was "DO NOT USE" in LIGO-T10005230-v1, so we moved it.

  2. Connected that ADC card to PEM channel box at 1X7 via SCSI cable.

  3. ADC card order is changed, so we checked ADC number assinging and re-labeled the cable.

  4. Found RFM is not working(c1sus and c1ioo not talking) and fb is in a weird state(Status: 0x4000 in GDS screens)

  5. Swapped the cabling so that ADC card 0 will be connected to timing interface card at slot1, but didn't help.
     More than that, we suffered ADC timeout.

  6. Tried ADC card swapping, slot position changing, taking out some of the ADC cards, etc.
     We found that ADC timeout doesn't happen with 2 ADC cards.
     But if we connect one of the ADC card to the timing interface card at slot 8, c1sus ADC timeouts with 2 ADC cards, too.
     So, I think that timing interface card is bad.

  7. Stopped rebooting c1sus again and again. We decided to investigate the problem tomorrow.
     We only need ADC card 0 and 1 for MC damping.(see this wiki page)
       ADC card 0: all UL/UR/LR/LL SENs
       ADC card 1: all SD SENs     
       ADC card 2: all QPDs

Result:
  We can damp optics and lock MC.
  We can't do A2L because RFM is not working.
  We can't see MC2 trans because we currently don't have ADC card 2.

  3945   Thu Nov 18 11:06:20 2010 josephbUpdateCDSc1sus and ADCs

Problem:

ADCs are timing out on c1sus when we have more than 3.

Talked with Rolf:

Alex will be back tomorrow (he took yesterday and today off), so I talked with Rolf.

He said ordering shouldn't make a difference and he's not sure why would be having a problem. However, when he loads the chassis, he tends to put all the ADCs on the same PCI bus (the back plane apparently contains multiples).  Slot 1 is its own bus, Slots 2-9 should be the same bus, and 10-17should be the same bus.

He also mentioned that when you use dmesg and see a line like "ADC TIMEOUT # ##### ######", the first number should be the ADC number, which is useful for determining which one is reporting back slow.

Plan:

Disconnect c1sus IO chassis completely, pull it out, pull out all cards, check connectors, and repopulate with Rolf's suggestions and keeping this elog in mind.

In regards to the RFM, it looks like one of the fibers had been disconnected from  the c1sus chassis RFM card (its plugged in in the middle of the chassis so its hard to see) during all the plugging in and out of the cables and cards last night.