  7537   Fri Oct 12 15:31:03 2012 RijuparnaConfiguration cavitymode scan

 Rijuparna, Manasa

Today I have checked the optical layout of the MC transmission RFPD table and measured the laser powers at different points. Manasa helped me for that. I found the power entering the RF photodiode is 0.394mW while the transmitted power of the cavity is 2.46mW. (I will give the diagram later).

  7051   Mon Jul 30 15:49:16 2012 steveUpdateVACcc1 switched

Quote:

Our cold cathode 423 gauges are 10 years old. They get insulated by age and show no ionization current.  We should replace them at the next vent. I'm buying 4 at $317ea

 

 We have two CC1 gauges at the pump spool. One is in vertical position and the spare is in horizontal. Today I switched the cable from vertical to horizontal CC1

This one is reading higher pressure. It is an old dirty gauge. It may trigger the RGA protection when it's reading goes above 1e-5 torr

The real pressure is 2e-6 torr

Attachment 1: cc1changedfVtH.png
cc1changedfVtH.png
  7056   Tue Jul 31 08:01:22 2012 steveUpdateVACcc1 switched

Quote:

Quote:

Our cold cathode 423 gauges are 10 years old. They get insulated by age and show no ionization current.  We should replace them at the next vent. I'm buying 4 at $317ea

 

 We have two CC1 gauges at the pump spool. One is in vertical position and the spare is in horizontal. Today I switched the cable from vertical to horizontal CC1

This one is reading higher pressure. It is an old dirty gauge. It may trigger the RGA protection when it's reading goes above 1e-5 torr

The real pressure is 2e-6 torr

 cc1h spikes closed VM1 last night. VM1 is opened.

Attachment 1: cc1h1d.png
cc1h1d.png
  6562   Tue Apr 24 14:55:00 2012 JamieUpdateCDScds code paths (rtscore, userapps) have moved

NOTE: Unless you really care about what's going on under the hood, please ignore this entire post and go here: USE THE NEW RTCDS UTILITY

Quote:

  • the userapps directory is in a new place (/opt/rtcds/userapps).  Not everything in the old location was checked into the repository, so we need to check to make sure everything that needs to be is checked in, and that all the models are running the right code.

 An important new aspect of this upgrade is that we have some new directories and some of the code paths have moved to correspond with the "standard" LIGO CDS filesystem hierarchy:

  • rtscore:      /opt/rtcds/rtscore/release       -->  RTS RCG source code (svn: adcLigoRTS)
  • userapps: /opt/rtcds/userapps/release  -->  CDS userapps source for models, RCG C code, medm screens, scripts, etc (svn: cds_user_apps)
  • rtbuild:       /opt/rtcds/caltech/c1/rtbuild    -->  RCG build directory

All work should be done in the "userapps" directory, and all builds should be done in the build directory.  Some important points:

WARNING: DO NOT MODIFY ANYTHING IN RTSCORE

This is important.  The rtscore directory is now just where the RCG source code is stored.  WE NO LONGER BUILD MODELS IN THE RTS SOURCE  Please use the rtbuild directory instead.

NO MORE MODEL/CODE SYMLINKS

You don't need to link you model or code anywhere to get it to compile.  The RCG now uses proper search paths to source in the RCG_LIB_PATH to find the needed source.  This has been configured by the admin, so as long as you put your code in the right place it should be found.

ALL CODE/MODELS/ETC GO IN USERAPPS

All RTS code is now stored ENTIRELY in the userapps directory (e.g. /opt/rtcds/userapps/release/isc/c1/models/c1lsc.mdl).  This is more-or-less the same as before, except that symlinking is no longer necessary.  I have placed a symlink at the old location for convenience.

BUILD ALL MODELS IN RTSCORE

You can run "make c1lsc" in the rtbuild directory, just as you used to in the old core directory.  However, don't do that.  USE THE NEW RTCDS UTILITY instead.

  9321   Thu Oct 31 00:06:45 2013 masayukiUpdateComputer Scripts / Programscds fft,tf,offsets

I wrote the scripts for cdsutils.

1. fft

usage: fft.py [-h] [-c N_CYCL] [-a N_AVG] freq channel [channel ...]

Measures the frequency content of one or more NDS channels
at the specified frequecy.
    The measurement results are  magnitude, phase, real part imaginary part, and the standard deviation of the real and imaginary parts.

positional arguments:
  freq        measurement frequency
  channel     list of measurement channel

optional arguments:
  -h, --help  show this help message and exit
  -c N_CYCL   define number of cycles. Default is 10
  -a N_AVG    define number of average. Default is 100

2. tf

usage: tf.py [-h] [-c N_CYCL] [-a N_AVG] [--dB] freq input output

Measures the transfer funtion of one NDS channels pair  at the specified frequecy.
    The measurement results are the coherence, magnitude, phase, real part, imaginary part, and the standard deviation of the real and imaginary parts.

positional arguments:
  freq        measurement frequency
  input       list of measurement input channel
  output      list of measurement output channel

optional arguments:
  -h, --help  show this help message and exit
  -c N_CYCL   define number of cycles. Default is 10
  -a N_AVG    define number of average. Default is 100
  --dB        print the amplitude in dB

 

3.offsets

usage: offset.py [-h] [-t SECONDS] filterbank [filterbank ...]

Zero the offset of one or more filterbank. Take average of OUT16 channel, and put (-1 * average / gain) into offset.  After change offsets, it will turn on the offset.
    example usage) offset -d 25 C1:LSC-AS55_Q C1:LSC-AS55_I

positional arguments:
  filterbank  list of filterbank to zero offset

optional arguments:
  -h, --help  show this help message and exit
  -t SECONDS  define the averaging time. default value is 10

I put these scripts in /opt/rtcds/cdsutils/trunk/lib/cdsutils.
I couldn't put them into cds library, but I will left tomorrow to Japan...

So please help me Jamie or Joe !!

 

 

By the way,

I modified the LSCoffset script  (script)/LSC/LSCoffsets.py.

usage: LSCoffsets.py [-h] [-d SECONDS] [--PDlist]

 Zero the offsets of LSC PDs. During taking average, we will close the PSL and green shutter. After zeroing, it will turn on the offsets switch.
    example usgae) LSCoffset2.py -d 20    

optional arguments:
  -h, --help  show this help message and exit
  -d SECONDS  define the averaging time. default value is 10
  --PDlist    print PD list for LSC and exit

i made new directory 'offset_backup' for old offset script. I move all old offset script into there.

But unfortunately that cannot use right now, because cds offsets script is not available yet...

  15905   Thu Mar 11 18:46:06 2021 gautamUpdateCDScds reboot

Since Koji was in the lab I decided to bite the bullet and do the reboot. I've modified the reboot script - now, it prompts the user to confirm that the time recognized by the FEs are the same (use the IOP model's status screen, the GPSTIME is updated live on the upper right hand corner). So you would do sudo date --set="Thu 11 Mar 2021 06:48:30 PM UTC" for example, and then restart the IOP model. Why is this necessary? Who knows. It seems to be a deterministic way of getting things back up and running for now so we have to live with it. I will note that this was not a problem between 2017 and 2020 Oct, in which time I've run the reboot script >10 times without needing to take this step. But things change (for an as of yet unknown reason) and we must adapt. Once the IOPs all report a green "DC" status light on the CDS overview screen, you can let the script take you the rest of the way again.

The main point of this work was to relax the data rate on the c1lsc model, and this worked. It now registers ~3.2 MB/s, down from the ~3.8 MB/s earlier today. I can now measure 2 loop TFs simultaneously. This means that we should avoid adding any more DQ channels to the c1lsc model (without some adjustment/downsampling of others).

Quote:

 Holding off on a restart until I decide I have the energy to recover the CDS system from the inevitable crash.

Attachment 1: CDSoverview.png
CDSoverview.png
  15910   Fri Mar 12 03:28:51 2021 KojiUpdateCDScds reboot

I want to emphasize the followings:

  • FB's RTC (real-time clock/motherboard clock) is synchronized to NTP time.
  • The other RT machines are not synchronized to NTP time.
  • My speculation is that the RT machine timestamp is produced from the local RT machine time because there is no IRIG-B signal provided. -> If the RTC is more than 1 sec off from the FB time, the timestamp inconsistency occurs. Then it induces "DC" error.
  • For now, we have to use "date" command to match the local RTC time to the FB's time.
     
  • So: If NTP synchronization is properly implemented for the RT machines, we will be free from this silly manual time adjustment.

 

  15911   Fri Mar 12 11:02:38 2021 gautamUpdateCDScds reboot

I looked into this a bit today morning. I forgot exactly what time we restarted the machines, but looking at the timesyncd logs, it appears that the NTP synchronization is in fact working (log below is for c1sus, similar on other FEs):

-- Logs begin at Fri 2021-03-12 02:01:34 UTC, end at Fri 2021-03-12 19:01:55 UTC. --
Mar 12 02:01:36 c1sus systemd[1]: Starting Network Time Synchronization...
Mar 12 02:01:37 c1sus systemd-timesyncd[243]: Using NTP server 192.168.113.201:123 (ntpserver).
Mar 12 02:01:37 c1sus systemd[1]: Started Network Time Synchronization.
Mar 12 02:02:09 c1sus systemd-timesyncd[243]: Using NTP server 192.168.113.201:123 (ntpserver).

So, the service is doing what it is supposed to (using the FB, 192.168.113.201, as the ntpserver). You can see that the timesync was done a couple of seconds after the machine booted up (validated against "uptime"). Then, the service is periodically correcting drifts. idk what it means that the time wasn't in sync when we check the time using timedatectl or similar. Anyway, like I said, I have successfully rebooted all the FEs without having to do this weird time adjustment >10 times

I guess what I am saying is, I don't know what action is necessary for "implementing NTP synchronization properly", since the diagnostic logfile seems to indicate that it is doing what it is supposed to.

More worryingly, the time has already drifted in < 24 hours.

Quote:

I want to emphasize the followings:

  • FB's RTC (real-time clock/motherboard clock) is synchronized to NTP time.
  • The other RT machines are not synchronized to NTP time.
  • My speculation is that the RT machine timestamp is produced from the local RT machine time because there is no IRIG-B signal provided. -> If the RTC is more than 1 sec off from the FB time, the timestamp inconsistency occurs. Then it induces "DC" error.
  • For now, we have to use "date" command to match the local RTC time to the FB's time.
     
  • So: If NTP synchronization is properly implemented for the RT machines, we will be free from this silly manual time adjustment.
Attachment 1: timeDrift.png
timeDrift.png
  11130   Tue Mar 10 20:08:23 2015 ericqUpdateCDScdsRampMuxMatrix Backported to 40m RCG

I have successfully backported the cdsRampMuxMatrix part for use in our RCG v2.5 system. This involved grabbing new files, merging changes, and hacking around missing features from RCG 2.9. 

The added/changed files, with relative paths referred to /opt/rtcds/rtscore/release/src/, are:

  • M include/drv/tRamp.c
    • New C functions to directly report current ramping value and ramping state 
  • M epics/util/feCodeGen.pl
    • Added if statements to main simulink parser to properly handle the part
  • AM epics/util/lib/RampMuxMatrix.pm
    • New Perl script that writes the frontend code for a given ramp matrix
  • A epics/util/mkrampmatrix.pl
    • New perl script that creates the default MEDM screen
  • A epics/simLink/lib/cdsRampMuxMatrix.mdl
    • Simulink block for the part
  • M epics/simLink/CDS_PARTS.mdl
    • Added block and doc for the part (which is missing an underscore in its definition of EPICS field names)

[A means the file was added, M means the file was modified]

Most of the trouble came from the EPICS reporting of the live ramping value and ramping state, since this depended on some future RCG value masking function. I had to rewrite the C-code writing perl script to define and update these EPICS variables in a more old-school way. 

This leaves us vunerable to the fact that a user/program can directly write to the live matrix element and ramping state, which would cause bad and unexpected behavior of the matrix.

So, when using a ramping matrix: NEVER WRITE to [MAT]_[N]_[N] as you would for a normal matrix. Use [MAT]_SETTING_[N]_[M] and trigger [MAT]_LOAD_MATRIX.

Similary, [MAT]_RAMPING_[N]_[N] is off limits. 

I tested the new part in the c1tst model. There are two EPICS input (TST-RAMP1_IN and TST-RAMP2_IN) that are the inputs to a 2x2 ramp matrix called TST-RAMP, and the outputs go to two testpoints (TST-RAMP1_OUT and TST-RAMP2_OUT) and two epics outputs (TST-RAMP1_OUTMON and TST-RAMP2_OUTMON). You can write something to the inputs from ezca or whatever, and use the C1TST_RAMP.adl medm screen in the c1tst directory to try it out. The buttons turn red when you've input a new matrix, yellow when a ramp is ongoing and green when the live value agrees with the setting. 

At this time, I have not rebuilt any of our operational models in search of potential issues.

I have created backups of the files I modified, such that a file such as feCodeGen.pl was renamed feCodeGen.40m.pl, and left next to the modified file. I am open to more robust ways of doing the backup; since our RCG source is an svn checkout of the v2.5 branch (with local modifications, to boot), I suppose we don't want to commit there. Maybe we make a 40m branch? A seperate repo? 

  9075   Tue Aug 27 19:50:06 2013 JamieConfigurationComputer Scripts / Programscdsutils checked out into /opt/rtcds

I have checked out the new cdsutils repository at:

/opt/rtcds/cdsutils/release

This is a new repository that is intended to hold all of our python libraries and command-line utilities for interacting with the IFO, things like:

  • get/write values EPICS channels
  • interact with filter module switches
  • average a test point for some amount of time
  • etc.

Basically everything that used to be ez* or tds*.

There's not much in there at the moment, but hopefully it will start to get filled in soon.

WARNING:

This code in here will be used by the sites to interact with the real aLIGO IFOs.  Please be careful as you develop things in here, and o so conscientiously.  If you do bad things here and it messes things up at the sites people will be angry.  Particularly me, since I have to support everything in here for Guardian use.

Usage

<cdsutils>/lib/cdsutils is the primary python library.  For each function you want to add, put it in a new file named after the function.  So for instance function "foo" should be in a file called <cdsutils>/lib/cdsutils/foo.py.

There is a command line utility at <cdsutils>/bin/cdsutils.  It will automatically find anything you add to the library and expose it as a sub command (e.g. "cdsutils foo")

We'll try to put together a wiki page describing development and usage of this soon.

  9132   Mon Sep 16 15:29:50 2013 JamieConfigurationComputer Scripts / Programscdsutils checked out into /opt/rtcds

We now have a proper install of cdsutils:

 controls@rossa:~ 0$ cdsutils
 usage: cdsutils <cmd> <args>

 Advanced LIGO Control Room Utilites

 Available commands:

   read         read EPICS channel value
   write        write EPICS channel value
   switch       switch buttons in standard LIGO filter module
   avg          average NDS channels for some amount of time
   servo        simple integrator (pole at zero)

 Add '-h' after individual commands for command help.
 controls@rossa:~ 0$ 

It is installed in /ligo/apps/cdsutils, and should be in the path on all workstations.

The "development" source working directory is currently checked out at /opt/rtcds/cdsutils/trunk.

 

  9133   Mon Sep 16 19:41:01 2013 ranaConfigurationComputer Scripts / Programscdsutils checked out into /opt/rtcds
 controls@rosalba:~ 0$ cdsutils
Traceback (most recent call last):
  File "/ligo/apps/cdsutils/lib/cdsutils/__main__.py", line 7, in <module>
    from cdsutils import CMDS
  File "/ligo/apps/cdsutils/lib/cdsutils/__init__.py", line 4, in <module>
    from servo import servo
  File "/ligo/apps/cdsutils/lib/cdsutils/servo.py", line 1, in <module>
    from epics import PV
ImportError: No module named epics
controls@rosalba:~ 1$
Mon Sep 16 19:40:32 2013 
  10040   Sun Jun 15 14:26:30 2014 JamieOmnistructureCDScdsutils re-installed

Quote:

 CDSUTILS is also gone from the path on all the workstations, so we need Jamie to tell us by ELOG how to set it up, or else we have to use ezcaread / ezcawrite forever.

It's in the elog already: http://nodus.ligo.caltech.edu:8080/40m/9922

But it seems like things still haven't fully recovered, or have recovered to an old state?  Why is the cdsutils install I previously did in /ligo/apps now missing?  It seems like other directories are missing as well.

There's also a user:group issue with the /home/cds mounts.  Everything in those mount points is owned nobody:nogroup.

I also can't log into pianosa and rosalba.

  10041   Sun Jun 15 14:41:08 2014 JamieOmnistructureCDScdsutils re-installed

Quote:

Quote:

 CDSUTILS is also gone from the path on all the workstations, so we need Jamie to tell us by ELOG how to set it up, or else we have to use ezcaread / ezcawrite forever.

It's in the elog already: http://nodus.ligo.caltech.edu:8080/40m/9922

But it seems like things still haven't fully recovered, or have recovered to an old state?  Why is the cdsutils install I previously did in /ligo/apps now missing?  It seems like other directories are missing as well.

There's also a user:group issue with the /home/cds mounts.  Everything in those mount points is owned nobody:nogroup.

I also can't log into pianosa and rosalba.

 I also still think it's a bad idea for everything to be mounting /home/cds from an IP address.  Just make a new DNS entry for linux1 and leave everything as it was.

  10057   Wed Jun 18 13:39:01 2014 ranaUpdateCDScdsutils reverted to version 238

 

 After some email consult with Jamie, I got cdsutils working again by reverting to v238. The newest versions are not compatible with our python 2.6 on Ubuntu 10. And our Ubuntu 12 machines do not have NDS2 clients that work yet.

The read/write commands work at the moment, but the NDS based ones don't yet work on pianosa due to some NDSSERVER flag/setup issue maybe, Jamie ??

controls@pianosa|~ > z

usage: cdsutils <cmd> <args>

Advanced LIGO Control Room Utilites

Available commands:

    read         read EPICS channel value

  write        write EPICS channel value

  switch       switch buttons in standard LIGO filter module

  avg          average one or more NDS channels for a specified amount of seconds

  servo        servos channel with a simple integrator (pole at zero)

  trigservo    servos channel with a simple integrator (pole at zero)


  version      print version info and exit

  help         this help


Add '-h' after individual commands for command help.

controls@pianosa|~ > z read C1:LSC-DARM_GAIN

0.0

controls@pianosa|~ 2> z avg 3 C1:IOO-MC_F

Error in write(): Connection refused

Error in write(): Connection refused

NDSSERVER variable incorrectly defined, or no NDS servers available.

controls@pianosa|~ 1> echo $NDSSERVER

fb

  9926   Wed May 7 23:30:21 2014 jamieUpdateCDScdsutils should be working now

Should be fixed now.  There were python2.6 compatibility issues, which only show up on these old distros (e.g. ubuntu 10.04).

controls@pianosa:~ 0$ cdsutils read C1:LSC-DARM_GAIN
0.0
controls@pianosa:~ 0$ cdsutils --version
cdsutils 230
controls@pianosa:~ 0$ 
  9922   Wed May 7 16:31:12 2014 jamieUpdateCDScdsutils updated to version 226
controls@pianosa:~ 0$ cd /opt/rtcds/cdsutils/trunk/
controls@pianosa:/opt/rtcds/cdsutils/trunk 0$ svn update
...
At revision 226.
controls@pianosa:/opt/rtcds/cdsutils/trunk 0$ make
echo "__version__ = '226'" >lib/cdsutils/_version.py
echo "__version__ = '226'" >lib/ezca/_version.py
...
controls@pianosa:/opt/rtcds/cdsutils/trunk 0$ make ligo-install
python ./setup.py install --prefix=/ligo/apps/linux-x86_64/cdsutils-226
...
controls@pianosa:/opt/rtcds/cdsutils/trunk 0$ ln -sfn cdsutils-226 /ligo/apps/linux-x86_64/cdsutils
controls@pianosa:/opt/rtcds/cdsutils/trunk 0$ exit
...
controls@pianosa:~ 0$ cdsutils --version
cdsutils 226
controls@pianosa:~ 0$ 

  9923   Wed May 7 17:10:59 2014 ranaUpdateCDScdsutils updated to version 226

 This upgrade from Jamie has given us the new apps (avg, servo, and trigservo). We should figure out if there's a way to integrate Masayuki's work, so that we can have a 'cdsutils demod' function too.

  9924   Wed May 7 22:47:33 2014 ranaUpdateCDScdsutils updated to version 226: not working on pianosa or rossa

 controls@rossa:~ 0$ cdsutils read C1:LSC-DARM_GAIN
Traceback (most recent call last):
  File "/usr/lib/python2.6/runpy.py", line 122, in _run_module_as_main
    "__main__", fname, loader, pkg_name)
  File "/usr/lib/python2.6/runpy.py", line 34, in _run_code
    exec code in run_globals
  File "/ligo/apps/cdsutils-226/lib/python2.6/site-packages/cdsutils/__main__.py", line 57, in <module>
    imp.load_module('__main__', f, pathname, description)
  File "/ligo/apps/cdsutils-226/lib/python2.6/site-packages/cdsutils/read.py", line 32, in <module>
    print ezca.Ezca(prefix).read(rest, as_string=args.as_string)
  File "/ligo/apps/linux-x86_64/cdsutils-226/lib/python2.6/site-packages/ezca/cached.py", line 17, in __call__
    key = (args, tuple(kwargs.viewitems()))
AttributeError: 'dict' object has no attribute 'viewitems'

  10188   Fri Jul 11 22:02:52 2014 Jamie, ChrisOmnistructureCDScdsutils: multifarious upgrades

To make the latest cdsutils available in the control room, we've done the following:

Upgrade pianosa to Ubuntu 12 (cdsutils depends on python2.7, not found in the previous release)

  • Enable distribution upgrades in the Ubuntu Software Center prefs
  • Check for updates in the Update Manager and click the big "Upgrade" button

Note that rossa remains on Ubuntu 10 for now.

Upgrade cdsutils to r260

  • Instructions here
  • cdsutils-238 was left as the default pointed to by the cdsutils symlink, for rossa's sake

Built and installed the nds2-client (a cdsutils dependency)

  • Checked out the source tree from svn into /ligo/svncommon/nds2
  • Built tags/nds2_client_0_10_5 (install instructions are here; build dependencies were installed by apt-get on chiara)
  • ./configure --prefix=/ligo/apps/ubuntu12/nds2-client-0.10.5; make; make install
  • In /ligo/apps/ubuntu12: ln -s nds2-client-0.10.5 nds2-client

nds2-client was apparently installed locally as a deb in the past, but the version in lscsoft seems broken currently (unknown symbols?). We should revisit this.

Built and installed pyepics (a cdsutils dependency)

  • Download link to ~/src on chiara
  • python setup.py build; python setup.py install --prefix=/ligo/apps/ubuntu12/pyepics-3.2.3
  • In /ligo/apps/ubuntu12: ln -s pyepics-3.2.3 pyepics

pyepics was also installed as deb before; should revisit when Jamie gets back.

Added the gqrx ppa and installed gnuradio (dependency of the waterfall plotter)

Added a test in /ligo/apps/ligoapps-user-env.sh to load the new cdsutils only on Ubuntu 12.

The end result:

controls@chiara|~ > z
usage: cdsutils  

Advanced LIGO Control Room Utilites

Available commands:

  read         read EPICS channel value
  write        write EPICS channel value
  switch       switch buttons in standard LIGO filter module
  avg          average one or more NDS channels for a specified amount of seconds
  servo        servos channel with a simple integrator (pole at zero)
  trigservo    servos channel with a simple integrator (pole at zero)
  audio        Play channel as audio stream.
  dv           Plot time series of channels from NDS.
  water        Live waterfall plotter for LIGO data

  version      print version info and exit
  help         this help

Add '-h' after individual commands for command help.
 
 
  10339   Wed Aug 6 13:17:21 2014 ericqOmnistructureCDScdsutils: multifarious upgrades

I've checked out cdsutils-274 to /opt/rtcds/cdsutils, and updated the /ligo/apps/ligoapps-user-env.sh to have the newer machines use it by default. This was to gain access to the cdsutils.Step methods for use in the smooth ASS handoffs script. 

  3700   Tue Oct 12 22:06:08 2010 yutaUpdateComputerscelan-installed CentOS 5.5 on mafalda

I clean-installed CentOS 5.5(32bit) on mafalda.
No firewalls, no SELinix.

  3701   Tue Oct 12 23:35:12 2010 mafaldaUpdateComputerscelan-installed CentOS 5.5 on mafalda

Quote:

I clean-installed CentOS 5.5(32bit) on mafalda.
No firewalls, no SELinix.

 Yuta has removed my ethernet connection. Help me!!!

rossa:mDV>ping mafalda
PING mafalda.martian (192.168.113.23) 56(84) bytes of data.
From rossa.martian (192.168.113.215) icmp_seq=2 Destination Host Unreachable
From rossa.martian (192.168.113.215) icmp_seq=3 Destination Host Unreachable
From rossa.martian (192.168.113.215) icmp_seq=4 Destination Host Unreachable

--- mafalda.martian ping statistics ---
5 packets transmitted, 0 received, +3 errors, 100% packet loss, time 3999ms
, pipe 3

  10461   Fri Sep 5 16:25:36 2014 SteveUpdateSUScenter ITMY oplev

Quote:

The SRM qpd was moved to accommodate the HeNe laser qualification test for LIGO Oplev use.

The qpd was saturating at 65,000 counts of 3 mW

ND1 filter lowering the power by 10 got rid of saturation. I epoxied an adapter ring to the qpd.

Atm3 was taken before saturation was realized with Koji's help.

Atm4 ND1 on SRM qpd. Now it is working and everything is moving.

 

 ITMY oplev should be centered. I worked too much around it.

  6700   Tue May 29 10:08:21 2012 steveUpdateIOOcentered IOO monitoring qpds

IOO Angle & IOO Position QPDs centered.

Attachment 1: 40dtrendIOO.png
40dtrendIOO.png
Attachment 2: IOOcentered.png
IOOcentered.png
  9347   Tue Nov 5 17:12:34 2013 SteveUpdateLSCcentered qpds

 Full list tomorrow: IP-Ang & Pos, ETMY-T, ETMY-Oplev, ETMX-T, IOO-Ang & Pos

 RA: No one in the control room this evening can understand what this ELOG means. Please use more words.

Attachment 1: recentered.png
recentered.png
  5233   Sun Aug 14 20:04:40 2011 Keiko, Anamaria, Jenne, and KiwamuSummaryLockingcentral part ifo locking plan
GOAL : To lock the central part of ifo

Here is the plan:

Mon - assemble all the cables from PDs and mixers, and check the CDS channels. Prepare the beamsplitters.

Tue - The current paths to REFL11 and REFL55 will be modified to the four paths to REFL11, 33, 55, 165. And the PDs will be placed.
Wed, Thu - during waiting for the ifo available with vacuum, help aligning the POP, POX, POY. In parallel, a simulation to find the PRC length SRC 
length tolerance will be proceeded.

Fri - When the ifo becomes available with vacuum, the sensing signals by 3-f scheme will be obtained with proper demodulation phases.

Sat - Try to lock the central part of the ifo with the new 3-f signals.
  5243   Mon Aug 15 21:43:29 2011 Anamaria and KeikoSummaryLockingcentral part ifo locking project

 REFL33 and REFL165 cables were connected from the AP table to the rack.  Cables on the rack for REFL33I, 33Q, 165I, 165Q ports were connected, too. Connections were confirmed by the data viewer. Two SMA cables which will be used for the two PDs on the AP tabl were built. We will be able to place the two PDs tomorrow. The beamsplitters to split the laser to REFL33 and REFL165 ports were mounted and ready to be placed.

  7327   Fri Aug 31 10:24:36 2012 SteveUpdateVACchamber dog clamps checked

Quote:

I tightened as many of the dog clamps on the bottom of the BS, ITMX and ITMY chambers as I could find.  I used a torque wrench at 45 ft-lbs.  Some of the bolts of the dogs were too long, and I couldn't find an extender thing to accommodate the bolt so I could reach the nut.  None of the bolts moved that I was able to reach.

Steve, we're not doing final final alignment today (we will do it tomorrow), so please go around and double-check my work by checking all of the dogs first thing in the morning.  Thanks.

 Almost all chamber dog clamps on the floor checked. There are a few exception where it is impossible to to get to the nut. 

Only the OOC nuts turned little bit. So our elastomer discs are holding up well. This means that the chamber anchoring to the floor is good.

  7373   Wed Sep 12 08:16:49 2012 SteveConfigurationPEMchamber must be sealed overnight!

Quote:

We conducted a beam scan on the AP table of the AS beam. We used a lens to focus the beam onto a power meter, and slowly moved a razor blade across the beam using a micrometer, vertically and horizontally both in front of and behind the beam. We also had to block the beam next to the AS beam in order to do this, but is unblocked now. Mike will begin curve fitting the data to try and see if there is a different spot size given by the x-axis vs. the y-axis, and if the lens has any effect.

 The vacuum envelope must be sealed with light doors on o-rings to insure a bug free IFO.  This was a violation!

  5975   Tue Nov 22 04:02:47 2011 kiwamuUpdateIOOchanged MC alignment

I have changed the MC2_YAW DC bias because the PZT1_YAW was railing.

I also realigned the steering mirrors in zig-zag path since the mode cleaner tended to resonate with higher order modes after I have changed the MC2 bias.

  C1:SUS-MC2_YAW_COMM =  -1.1548    => -1.1208

  7459   Mon Oct 1 19:21:03 2012 ranaConfigurationPEMchanged PEM DQ channels

Changed the list of channels to be written to frames from having the IN1 suffix to OUT. Now we can load the calibration of the channel into the filter module and the DQ channel will be calibrated.

We should do this wherever possible so that our channels will have real calibrations associated with them.
SEIS_GUR1_X_OUT 256
SEIS_GUR1_Y_OUT 256
SEIS_GUR1_Z_OUT 256
SEIS_GUR2_X_OUT 256
SEIS_GUR2_Y_OUT 256
SEIS_GUR2_Z_OUT 256
SEIS_STS_1_X_OUT 256
SEIS_STS_1_Y_OUT 256
SEIS_STS_1_Z_OUT 256
SEIS_STS_2_X_OUT 256
SEIS_STS_2_Y_OUT 256
SEIS_STS_2_Z_OUT 256
SEIS_STS_3_X_OUT 256
SEIS_STS_3_Y_OUT 256
SEIS_STS_3_Z_OUT 256
MIC_1_OUT 2048
MIC_2_OUT 2048
MIC_3_OUT 2048
MIC_4_OUT 2048
MIC_5_OUT 2048
MIC_6_OUT 2048
ACC_MC1_X_OUT 2048
ACC_MC1_Y_OUT 2048
ACC_MC1_Z_OUT 2048
ACC_MC2_X_OUT 2048
ACC_MC2_Y_OUT 2048
ACC_MC2_Z_OUT 2048
XARM_DIFFERENTIAL_MOTION_IN1 256
XARM_DIFFERENTIAL_MOTION_OUT 256
YARM_DIFFERENTIAL_MOTION_IN1 256
YARM_DIFFERENTIAL_MOTION_OUT 256

Next we should up the rate at which the model runs up to 16 kHz so that we can record the microphones at 16 kHz. FM radio has information up to 20 kHz. AM radio goes up to ~8 kHz. We should be at least as modern as AM radio. How do we make the change? How do we make sure the FOTON file stays OK?

I have made some changes to the daily summary file to compensate. New files is /users/public_html/40m-summary/share/c1_summary_page.ini.

  7488   Fri Oct 5 01:36:49 2012 DenConfigurationPEMchanged PEM DQ channels

Quote:

We should do this wherever possible so that our channels will have real calibrations associated with them.

Next we should up the rate at which the model runs up to 16 kHz so that we can record the microphones at 16 kHz. FM radio has information up to 20 kHz. AM radio goes up to ~8 kHz. We should be at least as modern as AM radio. How do we make the change? How do we make sure the FOTON file stays OK?

 I've added calibration gains to Guralp (to um/sec) and EM172 (to Pa) channels.

We can run PEM at 16 kHz. I think Foton file stores both sos-representation and filter commands which are independent of the sampling frequency, so it should be possible to change model sampling frequency quickly.

In fact, we can save data at 64 kHz from iop models. I've done this once with MC_F channel. However, I did not test EM172 noise at frequencies > 1 kHz.

  5397   Tue Sep 13 19:12:53 2011 SureshUpdateComputer Scripts / Programschanges to the c1ioo, c1mcs and c1rfm

[Suresh, Jamie]

The new versions of these three models have been committed to the svn.  I copy below the svn log I wrote over there.

I edited three models,namely c1ioo, c1mcs and c1rfm to bring four channels from C1MCS model into the
 C1IOO model through the RFM.

We plan to use the MC2-Trans-QPD as a third sensor in addition to the WFS1 and WFS2 in the ASC system so that we can sense all the six degrees of freedom of the MC.  However the MC2-Trans_QPD is a new incarnation of the old MC2 Oplev QPD and its four segments are are picked up in the c1mcs model running on the C1SUS machine.  Since we need these    in the C1IOO machine we    had to port these signals via the RFM.  The    changes    I made to these    models are described below:
1) c1rfm.mdl -->  I added several cds parts to carry these four    channels from c1mcs to c1ioo via c1rfm

2) c1mcs.adl -->  In the MC2 block of c1mcs I cut out the structure which converts the four segments of the  QPD in the pitch, yaw and sum signals and copied it to the c1ioo model.     In the    c1mcs I  grounded the input of the oplev signals just as they are in    the MC1    and MC2    suspension blocks. I then routed the ADC channels associated with the oplev QPD signals to the rfm model.

3) c1ioo.mdl --> I have made extensive changes to the c1ioo model to include the third sensor in the WFS feedback loop.  The six lockins in the model are used to excite the MC mirrors and the signals from WFS and QPD can be demodulated to determine the input or output matrices in the WFS feedback.  A sensor_demod_array permits us to select the signals we wish to demodulate. The MCL signal, used in the MC_ASS and A2L measurements, is retained as one of the signals.  The MC_TRANS_SUM was used several in  scripts and to preserve the continuity with this channel name we inserted three EPICS parts to generate a slow channel by the same name from this model.

Since the MC_TRANS_SUM is used in the MC-autolocker script we checked to make sure that the script is running after we recompiled all the three models and restarted them.  It is all fine.

  1982   Thu Sep 10 17:47:25 2009 JenneUpdateComputerschanges to the startass scripts

[Rana, Jenne]

While I was mostly able to restart the c1ass computer earlier today, the filter banks were acting totally weird.  They were showing input excitations when we weren't putting any, and they were showing that the outputs were all zero, even though the inputs were non-zero and the input and the output were both enabled. The solution to this ended up being to use the 2nd to last assfe.rtl backup file.  Rana made a symbolic link from assfe.rtl to the 2nd to last backup, so that the startup.cmd script does not need to be changed whenever we alter the front end code.

The startup_ass script, in /caltech/target/gds/ which, among other things, starts the awgtpman was changed to match the instructions on the wiki Computer Restart page.  We now start up the /opt/gds/awgtpman .  This may or may not be a good idea though, since we are currently not able to get channels on DTT and Dataviewer for the C1:ASS-TOP_PEM channels.  When we try to run the awgtpman that the script used to try to start ( /caltech/target/gds/bin/ ) we get a "Floating Exception". We should figure this out though, because the /opt/gds/awgtpman does not let us choose 2kHz as an option, which is the rate that the ASS_TOP stuff seems to run at.

The last fix made was to the screen snapshot buttons on the C1:ASS_TOP screen.  When the screen was made, the buttons were copied from one of the other ASS screens, so the snapshots saved on the ASS_TOP screen were of the ASS_PIT screen.  Not so helpful.  Now the update snapshot button will actually update the ASS_TOP snapshot, and we can view past ASS_TOP shots.

  4637   Thu May 5 13:14:15 2011 kiwamuUpdateLSCchans file for LSC

Foton doesn't correctly display the LSC filter bank file : C1LSC.txt.

Foton tells a lie that they all are empty.

 

The file itself looks fine to me i.e. I can find correct filters in text format.

Looks like someone (maybe Joe and Jenne ?) updated the file. I am not sure if this is the reason or not.

allegra:chans>ls -al | grep LSC
-rw-r--r--  1 controls controls   20659 May  5 11:46 C1LSC.txt

 

NEEDS TO BE FIXED SOON

  4640   Thu May 5 14:50:30 2011 JenneUpdateLSCchans file for LSC

Quote:

Foton doesn't correctly display the LSC filter bank file : C1LSC.txt.

Foton tells a lie that they all are empty.

 

The file itself looks fine to me i.e. I can find correct filters in text format.

Looks like someone (maybe Joe and Jenne ?) updated the file. I am not sure if this is the reason or not.

allegra:chans>ls -al | grep LSC
-rw-r--r--  1 controls controls   20659 May  5 11:46 C1LSC.txt

 

NEEDS TO BE FIXED SOON

 See my updated elog 4636 for what Joe and I did this morning, and what a possible problem is (making the LSC model into a sub-model).

  4645   Thu May 5 16:11:22 2011 JamieUpdateLSCchans file for LSC

Quote:

Foton doesn't correctly display the LSC filter bank file : C1LSC.txt.

 This was because of a bug in the RCG for foton filter module naming when top names is being used.  Rebuilding the LSC front-end model with top_names (which was needed to get around another bug in the RCG) broke the filter file.  I manually fixed the file, so it should work now.

  6940   Sun Jul 8 19:31:53 2012 yutaUpdateLockingcharacterizing LSC arm lock by ALS error signal

RMS of X/Y arm length using POX/POY lock is <160 pm and <120 pm respectively. RMS of free swinging X/Y arm length is both 0.17 um.

I used ALS error signal for out-of-loop evaluation of IR lock. We can even use ALS error signal when arm is free swinging because phase tracking ALS error signal is linear to arm length.
ALS error signal might not be as good as POX/POY. So, this out-of-loop estimation might be not so good.

X arm lock using POX11:
- Openloop transfer function
   I adjusted filter (C1:LSC-XARM) gain and now, UGF ~150 Hz, phase margin ~20 deg.
  570 usec delay (number in the figure is wrong) - Edited by Yuta on July 9
LSCPOXarmIRlockOLTF.png

- Arm length spectra
   Red is the free run spectrum. Measured using C1:ALS-BEATX_FINE_PHASE_OUT, calibration factor in frequency is 9.81 kHz/deg (see elog #6938), so calibration factor is 1.32 nm/deg.
   Green is the out-of-loop spectrum. Measured using C1:ALS-BEATX_FINE_PHASE_OUT.
   Blue is the in-loop spectrum. Measured using C1:LSC-POX11_I_ERR, calibration factor is 3.8e12 counts/m (see elog #6841).
   Black is the expected spectrum from openloop transfer function, such as (free run)/|1+G|.
XarmLengthspectra20120708.png


  Out-of-loop estimation of RMS during POX lock is 160 pm. But since this looks too large, ALS error signal might not see actual arm length change when arm length is locked.
  Also, it is interesting that ALS error signal sees 24 Hz peak, but POX doesn't. Roll mode coupling to green?

Y arm lock using POY11:
- Openloop transfer function
   I adjusted filter (C1:LSC-YARM) gain and now, UGF ~150 Hz, phase margin ~20 deg.
  570 usec delay (number in the figure is wrong) - Edited by Yuta on July 9
LSCPOYarmIRlockOLTF.png

- Arm length spectra
   Red is the free run spectrum. Measured using C1:ALS-BEATY_FINE_PHASE_OUT, calibration factor in frequency is 9.65 kHz/deg (see elog #6938), so calibration factor is 1.30 nm/deg.
   Green is the out-of-loop spectrum. Measured using C1:ALS-BEATY_FINE_PHASE_OUT.
   Blue is the in-loop spectrum. Measured using C1:LSC-POY11_I_ERR, calibration factor is 1.4e12 counts/m (see elog #6834).
   Black is the expected spectrum from openloop transferfunction, such as (free run)/|1+G|.
YarmLengthspectra20120708.png


  Out-of-loop estimation of RMS during POY lock is 120 pm. But since this looks too large, ALS error signal might not see actual arm length change when arm length is locked.
  Also, it is interesting that ALS error signal sees 16.5 Hz peak, but POY doesn't. Bounce mode coupling to green?

Next:
  - Noise budgeting of phase tracking ALS
  - Is it possible to lock MI when RMS of arm length during POX/POY lock increased to ~100pm?

  16121   Wed May 5 13:05:07 2021 ChubUpdateGeneralchassis delivery from De Leone

Assembled chassis from De Leone placed in the 40 Meter Lab, along the west wall and under the display pedestal table.  The leftover parts are in smaller Really Useful boxes, also on the parts pile along the west wall.

Attachment 1: de_leone_del_5-5-21.jpg
de_leone_del_5-5-21.jpg
  16160   Tue May 25 17:08:17 2021 ChubUpdateElectronicschassis rework complete!

All remaining chasses have been reworked and placed on the floor along the west wall in Room 104. 

Attachment 1: 40M_chassis_reworked_5-25-21.jpg
40M_chassis_reworked_5-25-21.jpg
  428   Fri Apr 18 19:46:08 2008 ranaUpdateASScheck adaptive
I restarted the adaptive code today using 'startass' and 'upass'.
I moved them into the scripts/ASS/ subdirectory.

Things seem OK. With a MU=0.03 and a TAU=0.00001, there is a still
a good factor of 10 reduction of the 3 Hz stack peak from the MC2
drive by doing FF into MC1.

I edited the ASS-TOP screen so that we could see such small numbers. I
also re-aligned the MC SUS to match the input beam (mainly MC3). The
cavity was locking on a TEM10 mode mostly -- we should look in the SUS
OSEM trends to see if MC3 has moved a lot in the last month or so.

Caryn Palatchi (a Caltech undergrad who just started working with us)
illustrated to me today that using even 1000 FIR taps is not very effective
for low frequency noise cancellation if you have a 2048 Hz sample rate. More
precisely, the asymptotic Wiener filter which our 'LMS' algorithm converges
to, can often amplify the noise at frequencies below f_sample/N_taps.

A less obvious thing that she also noticed is that there is almost no cancellation
of the 16.25 Hz bounce mode when using such a short filter. That's because that
mode is fairly high Q: the transfer function from the Z-ACC to the cavity signal
goes through the high-Q vertical suspension resonance; the FF signal we send back
goes through the low-Q horizontal pendulum response only. Therefore the filter
needs to be able to simulate ~100 cycles at 16.25 Hz in order to cancel that peak.

Duh.

The message here is: we need to find a computationally efficient way to do FIR filtering
or its not going to ever be cool enough to help us find the Crab.
Attachment 1: 0052_xray_thm45.jpg
0052_xray_thm45.jpg
  432   Mon Apr 21 12:58:42 2008 robUpdateASScheck adaptive

Quote:


Caryn Palatchi (a Caltech undergrad who just started working with us)
illustrated to me today that using even 1000 FIR taps is not very effective
for low frequency noise cancellation if you have a 2048 Hz sample rate. More
precisely, the asymptotic Wiener filter which our 'LMS' algorithm converges
to, can often amplify the noise at frequencies below f_sample/N_taps.

A less obvious thing that she also noticed is that there is almost no cancellation
of the 16.25 Hz bounce mode when using such a short filter. That's because that
mode is fairly high Q: the transfer function from the Z-ACC to the cavity signal
goes through the high-Q vertical suspension resonance; the FF signal we send back
goes through the low-Q horizontal pendulum response only. Therefore the filter
needs to be able to simulate ~100 cycles at 16.25 Hz in order to cancel that peak.

Duh.

The message here is: we need to find a computationally efficient way to do FIR filtering
or its not going to ever be cool enough to help us find the Crab.


This is the reason for "RDNSAMP" parameter in the ASS code. The FIR filtration is applied at the downsampled rate, not the machine rate. So, if RDNSAMP=32, the effective sampling rate of the FIR filter is 64Hz, and thus noise cancellation should be good down to 64Hz/1000, or 64mHz, and the filter has an impulse response time that extends to 15 secs. I'm not convinced the filter length is what's limiting the performance at the bounce mode, but I agree that a faster FIR implementation would be good.
  4327   Fri Feb 18 20:06:59 2011 kiwamuSummarySUScheck f2p function on ETMX

 The plot below shows how the f2p filters work.

At -2 min I turned on the f2p filters.

 f2p_ETMX.png

  5161   Wed Aug 10 00:11:39 2011 jamieUpdateSUScheck of input diagnolization of ETMX after OSEM tweaking

Suresh and I tweaked the OSEM angles in ETMX yesterday.  Last night the ETMs were left free swinging, and today I ran Rana's peakFit scripts on ETMX to check the input diagnolization:

ETMX.png

It's well inverted, but the matrix elements are not great:

       pit       yaw       pos       side      butt
UL    0.3466    0.4685    1.6092    0.3107    1.0428
UR    0.2630   -1.5315    1.7894   -0.0706   -1.1859
LR   -1.7370   -1.5681    0.3908   -0.0964    0.9392
LL   -1.6534    0.4319    0.2106    0.2849   -0.8320
SD    1.0834   -2.6676   -0.9920    1.0000   -0.1101

The magnets are all pretty well centered in the OSEMS, and we worked at rotating the OSEMS such that the bounce mode was minimized.

Rana and Koji are working on ETMY now.  Maybe they'll come up with a better procedure.

  3807   Thu Oct 28 04:28:50 2010 yutaUpdateGreen Lockingchecked frequency counter SR620

(Kiwamu, Yuta)

Background:
  For green locking, we are planning to feedback frequency differential signal to ETM suspension for the final configuration.
  We don't have ETM suspension control system right now, so we are going to feedback the signal to X-end laser frequency for a test.
  We have two loops for the servo;
    1. coarse locking using frequency counter, feeding back to the laser temperature
    2. using VCO, feeding back to the laser PZT
  Today, we checked frequency counter SR620 and see how to get the small beat note signal(-48dBm; see elog #3771).

What we did:
  1. Using Marconi(RF signal generator), put RF signals to SR620 and see how small signal SR620 can see.
    It depends on the frequency. For 80MHz signal, you need more than about -9dBm.
       60MHz  >-12dBm
       70MHz  >-10dBm
       80MHz  >-9dBm
       90MHz  >-8dBm
      100MHz  >-7dBm

Since we are going to lock the frequency difference between X-end and PSL to 80MHz, we need at least +40dBm amp before putting the signal into SR620.

RF amplifier ZHL-32A has around +28dBm +28dB gain at 80MHz, so we need 2 of them.

  2. Marconi -> ZHL-32A -> ZHL-32A -> SR620 and see how small 80MHz signal SR620 can see.
    Around -68dBm. This should be enough.

  3. SR620 has "STRIP CHART" output on the rear panel. The output voltage is proportional to the mean frequency of the input.
    The output range is 0-8V. So in order to get 4V for 80MHz, set SCALE to 20MHz.

Plan:
 - find green beat again and see if SR620 can see it with double ZHL-32A configuration

  3809   Thu Oct 28 11:54:31 2010 KojiUpdateGreen Lockingchecked frequency counter SR620

ZHL-32A is a high power (well..., actually middle power) amplifier with the max output power of +29dBm (~1W!).
It seems to be overkill.
Your signal is so small so you don't need ZHL-32A, but can use small amp which we may have somewhere in the lab.

And the description:
"RF amplifier ZHL-32A has around +28dBm gain at 80MHz"
The unit is wrong.

Quote:

(Kiwamu, Yuta)

Background:
  For green locking, we are planning to feedback frequency differential signal to ETM suspension for the final configuration.
  We don't have ETM suspension control system right now, so we are going to feedback the signal to X-end laser frequency for a test.
  We have two loops for the servo;
    1. coarse locking using frequency counter, feeding back to the laser temperature
    2. using VCO, feeding back to the laser PZT
  Today, we checked frequency counter SR620 and see how to get the small beat note signal(-48dBm; see elog #3771).

What we did:
  1. Using Marconi(RF signal generator), put RF signals to SR620 and see how small signal SR620 can see.
    It depends on the frequency. For 80MHz signal, you need more than about -9dBm.
       60MHz  >-12dBm
       70MHz  >-10dBm
       80MHz  >-9dBm
       90MHz  >-8dBm
      100MHz  >-7dBm

Since we are going to lock the frequency difference between X-end and PSL to 80MHz, we need at least +40dBm amp before putting the signal into SR620.

RF amplifier ZHL-32A has around +28dBm gain at 80MHz, so we need 2 of them.

  2. Marconi -> ZHL-32A -> ZHL-32A -> SR620 and see how small 80MHz signal SR620 can see.
    Around -68dBm. This should be enough.

  3. SR620 has "STRIP CHART" output on the rear panel. The output voltage is proportional to the mean frequency of the input.
    The output range is 0-8V. So in order to get 4V for 80MHz, set SCALE to 20MHz.

Plan:
 - find green beat again and see if SR620 can see it with double ZHL-32A configuration

 

  3842   Mon Nov 1 23:31:05 2010 yutaUpdateCDSchecked input hardware filter in single frequency

Background:
  For input filter, we have analog whitening filter and also digital whitening filter. They have the same TF and when analog one is off, digital one should be on and vice versa.
  I made a python script that checks the switching automatically.

Method:
  Excite the suspension in a single frequency and see sensor inputs(XXSEN_IN1).
  Calculate the magnitude in the excitation frequency and compare it when digital whitening is off and on.
  When digital whitening is off, analog should be on, so sensor inputs should gone though the analog filter. That means the signal is multiplied by the TF of that filter, which makes the difference.

  We currently don't have excitation and I thought I have to wait, but instead of putting some extra excitation, I found that 60Hz line noise is useful.

Script:
  The script is /cvs/cds/caltech/users/yuta/scripts/WDWchecker.py
  For every sensor input, it;
    0. Stores current filter switching(XXSEN_SW1R)
    1. turns OFF the digital filter(FM1, using ezcaswitch)
    2. tdsdmd XXSEN_IN1 in 60Hz
    3. turns ON the digital filter
    4. tdsdmd XXSEN_IN1 in 60Hz
    5. divides mag(2.) by mag(4.) and calculate the analog filter gain in 60Hz
    6. Restores the filter switching in the state before the checking

Result:
  The results are;

C1:SUS-BS_ULSEN_IN1: 22.2 dB
C1:SUS-BS_URSEN_IN1: 18.7 dB
C1:SUS-BS_LRSEN_IN1: 22.7 dB
C1:SUS-BS_LLSEN_IN1: 16.0 dB
C1:SUS-BS_SDSEN_IN1: 21.5 dB
C1:SUS-ITMX_ULSEN_IN1: 16.9 dB
C1:SUS-ITMX_URSEN_IN1: 16.3 dB
C1:SUS-ITMX_LRSEN_IN1: 17.5 dB
C1:SUS-ITMX_LLSEN_IN1: 17.1 dB
C1:SUS-ITMX_SDSEN_IN1: 6.2 dB
C1:SUS-ITMY_ULSEN_IN1: 15.5 dB
C1:SUS-ITMY_URSEN_IN1: 16.5 dB
C1:SUS-ITMY_LRSEN_IN1: 17.4 dB
C1:SUS-ITMY_LLSEN_IN1: 16.3 dB
C1:SUS-ITMY_SDSEN_IN1: 18.0 dB
C1:SUS-PRM_ULSEN_IN1: 0.1 dB
C1:SUS-PRM_URSEN_IN1: 10.3 dB
C1:SUS-PRM_LRSEN_IN1: 13.1 dB
C1:SUS-PRM_LLSEN_IN1: -32.3 dB
C1:SUS-PRM_SDSEN_IN1: 14.6 dB
C1:SUS-SRM_ULSEN_IN1: 17.3 dB
C1:SUS-SRM_URSEN_IN1: 13.5 dB
C1:SUS-SRM_LRSEN_IN1: 1.6 dB
C1:SUS-SRM_LLSEN_IN1: 16.7 dB
C1:SUS-SRM_SDSEN_IN1: 18.3 dB

C1:SUS-MC1_ULSEN_IN1: 17.0 dB
C1:SUS-MC1_URSEN_IN1: 18.6 dB
C1:SUS-MC1_LRSEN_IN1: 14.9 dB
C1:SUS-MC1_LLSEN_IN1: 27.0 dB
C1:SUS-MC1_SDSEN_IN1: 16.6 dB
C1:SUS-MC2_ULSEN_IN1: 19.8 dB
C1:SUS-MC2_URSEN_IN1: 14.0 dB
C1:SUS-MC2_LRSEN_IN1: 20.8 dB
C1:SUS-MC2_LLSEN_IN1: 16.1 dB
C1:SUS-MC2_SDSEN_IN1: 17.3 dB
C1:SUS-MC3_ULSEN_IN1: 15.5 dB
C1:SUS-MC3_URSEN_IN1: 17.3 dB
C1:SUS-MC3_LRSEN_IN1: 18.2 dB
C1:SUS-MC3_LLSEN_IN1: 18.7 dB
C1:SUS-MC3_SDSEN_IN1: 16.8 dB


  Whitening filter has 18dB gain at 60Hz. (It's 3Hz pole, 30Hz zero, 100Hz zero and 0dB at DC)
  So, from the result, at least MC suspensions look like they have correct switching.
  But some channels doesn't look ok.
  We have to check those.

Plan:
   - check ITMX_SDSEN, PRM_ULSEN, PRM_LLSEN, SRM_LRSEN input filters
   - check the script and see if the script can really check. maybe the script needs some adjustments (# of averaging, multiple frequency, ......)
   - make AWG(, tdssine) work
   - check output hardware filter

By the way:
  fb is back. I don't know why. With help from Joe, I just compiled c1mcs again and again changing number of RFM channels.

  3675   Thu Oct 7 23:24:44 2010 yutaUpdateCDSchecking MC1 suspension damping

Background:
 The new CDS is currently being set up.
 We want to see if the damping servo of the suspensions are working correctly.
 But before that, we have to see if the sensors and the coils are working correctly.
 Among the 8 optics, MCs take top priority because of the green beam. for the alignment of the in-vac optics.

What I did:
 By seeing the 5 sensor outputs (C1:SUS-MC1_XXSEN_IN1, XX=UL,UR,LR,LL,SD) with the Data Viewer, I checked if all the coils are kicking in the supposed direction and all the sensors are sensing that kick correctly.

 All the matrices elements were set to the ideal values(-1 or 0 or 1) this time.

Result:
 They were perfect.
1. POSITION seemed to be POSITION
 When the offset(C1:SUS-MC1_SUSPOS_OFFSET) was added, all the sensor output moved to the same direction.
2. PITCH seemed to be PITCH
 When the offset(C1:SUS-MC1_PIT_COMM) was changed, UL&UR and LL&LR went to the different direction.
3. YAW seemed to be YAW
 When the offset(C1:SUS-MC1_YAW_COMM) was changed, UL&LL and UR&LR went to the different direction.
4. SIDE seemed to be SIDE
 When the offset(C1:SUS-MC1_SDSEN_OFFSET) was added, DC level of the SD sensor output changed.

Notes:

 c1mcs crashed many times during the investigation, and I had to kill and restart it again and again.
 It seemed to be easily crashed when filters are on, and so I couldn't check whether the damping servo is working correctly or not today.

Next work:

  - fix c1mcs (and maybe others)
  - check the damping servo by comparing the displacements of each 4 degrees of freedom when servo in off and on.

  3678   Fri Oct 8 12:21:11 2010 josephbUpdateCDSchecking MC1 suspension damping

Upon investigation, it appears that the c1mcs model was (and still is) timing out after a random amount of time. Or in other words, it at some point it was taking too long to do all the calculations for a single cycle and falling behind. The evidence for this is from the dmesg command when run on c1sus.

There's a bunch of lines like:

[ 8877.438002] c1mcs: cycle 568 time 62; adcWait 0; write1 0; write2 0; longest write2 0

[ 8877.438002] c1mcs: cycle 569 time 62; adcWait 0; write1 0; write2 0; longest write2 0

With a final line like: [ 8877.439003] c1mcs: ADC TIMEOUT 1 2405 37 2277

This last line indicates in fell so far behind it gave up.

However, this doesn't actually seem to be related to the amount of computation going on with the front end. I restarted the c1mcs model this morning by logging into the c1sus machine, and changing to the /opt/rtcds/caltech/c1/target/c1mcs/bin directory and running:

lsmod

sudo rmmod c1mcsfe

sudo insmod c1mcsfe.ko

The first line just lists the running modules. The second removes the c1mcs module, and the third starts it up again.

I proceeded to turn all the filters and and set all the matrix values while keeping an eye on the C1MCS-GDS_TP.adl screen and its timing indicator. It was running fine with all these turned on. Then I ran a dtt session from rosalba (going to /opt/apps/, typing bash, then source gds-env.bash, and finally diaggui) as well as a dataviewer and looked at 6 test point channels. It received data fine.

However, about 2 minutes after I had stopped doing this (although the dataviewer realtime session was still going) the USR timing jumped from about 20 microseconds to 35 microseconds, and the CPU Max timing jumped to the 50-60 microsecond range. At this point dmesg started reporting things like:

[54143.465613] c1mcs: cycle 1076 time 62; adcWait 0; write1 0; write2 0; longest write2 0

[54143.526004] c1mcs: cycle 2064 time 62; adcWait 0; write1 0; write2 0; longest write2 0

About a minute later the code gave up and reported a ADC timeout via dmesg. None of the other front ends seem to be affected.

I had to clear the test points manually after stopping dataviewer and dtt by going to rosalaba,using the sourced gds-env.bash, and running diag -l. I then typed "tp clear 36 *" to clear all the test points on the model with FEC number 36 (corresponding to c1mcs).

I removed and restarted c1mcs again, trying to turn on a few things at a time and letting it run for a few minutes to see if I could narrow down if its one particular filter perhaps reaching an underflow and starting to bog down the computations. However, after about 45 minutes of this, the model is still running and I've turned all the filters on and have been running about 8 test points with no problem, so the problem is clearly intermittent.

Quote:

Notes:
 c1mcs crashed many times during the investigation, and I had to kill and restart it again and again.
 It seemed to be easily crashed when filters are on, and so I couldn't check whether the damping servo is working correctly or not today.

Next work:

  - fix c1mcs (and maybe others)
  - check the damping servo by comparing the displacements of each 4 degrees of freedom when servo in off and on.

 

ELOG V3.1.3-