1991
|
Fri Sep 18 14:25:00 2009 |
rob | Omnistructure | PSL | water under the laser chiller |
rob, koji, steve
We noticed some water (about a cup) on the floor under the NESLAB chiller today. We put the chiller up on blocks and took off the side panel for a cursory inspection, but found no obvious leaks. We'll keep an eye on it. |
1992
|
Fri Sep 18 16:05:08 2009 |
Jenne | Omnistructure | PSL | water under the laser chiller |
Quote: |
rob, koji, steve
We noticed some water (about a cup) on the floor under the NESLAB chiller today. We put the chiller up on blocks and took off the side panel for a cursory inspection, but found no obvious leaks. We'll keep an eye on it.
|
The culprit has been found: One of the bottles of chiller water had a tiny leak in it, and apparently the floor is sloped just right to make it look like the water had been coming from under the chiller. All is well again in the world of chilled water. |
1993
|
Fri Sep 18 16:26:02 2009 |
steve | Summary | PSL | Neslab chiller is OK |
Rob found puddles of water very close to the chiller during lunch time. We raised the unit and took the side cover off. All surfaces were dry and the water level in the tub normal.
Later on we discovered that one of the Vons distilled water bottle was leaking. Jenne and I checked for excess amount of condensing water droplets inside the MOPA box.
On the bare,not insulated tubing and valve are loaded with droplets of water. Relative humidity is 44% at 24 C and HEPA filter speed set to 80 V in the enclosure.
|
1995
|
Wed Sep 23 19:36:41 2009 |
rana | Update | PSL | RC temperature performance |
This first plot shows the RC temperature channels' performance from 40 days ago, before we disabled the MINCO PID controller. Although RCTEMP is supposed to be the out of loop sensor, what we really care about is the cavity length and so I've plotted the SLOW. To get the SLOW on the same scale, I've multiplied the channel by 10 and then adjusted the offset to get it on the same scale.
The second plot shows a period after that where there is no temperature control of the can at all. Same gain scaling has been applied to SLOW as above, so that instead of the usual 1 GHz/V this plot shows it in 0.1 GHz/V.
The third plot shows it after the new PID was setup.
Summary: Even though the PID loop has more gain, the true limit to the daily fluctuations in the cavity temperature and the laser frequency are due to the in-loop sensors measuring the wrong thing. i.e. the out-of-loop temperature is too different from the in-loop sensor. This can possibly be cured with better foam and better placement of the temperature sensors. Its possible that we're now just limited by the temperature gradients on the can. |
2008
|
Sun Sep 27 14:45:45 2009 |
Koji | Update | PSL | SLOWscan result |
I ran (script dir)/PSL/FSS/SLOWscan on op440m from 11:30 to 12:30 on 27th. Although Rana and later I myself set "timed bombs" for the scan, they did not work as they have probably been ran on Linux. After the scan I relocked PMC, FSS, and MZ . MC locked automatically.
Observation:
1. To keep away from the mode hop, FSS_SLOWDC is to be at around 0. The values -5 ~ -6 is the place for the power, which is my preference for now. BTW, the mode hop only appears to the PSL output (=AMPMON) is this normal?
2. The PSL output looks dependent on the NPRO wavelength. The NPRO output and the PSL output tends to be high when the FSS_SLOWDC is low (= LTMP: Laser Crystal Temp is low). Also there is a step at the LTMP where we think the mode hop is present. This may cause the daily PSL output variation which induced by the daily change of the reference cavity length.
My naive speculation is that the NPRO wavelength is too long (= hot side) for the MOPA absorption as the MOPA heads are cooled to 19deg.
3. Scanning of -10 to +10 changes the LTMP from 42-49deg. This is almost 1/10 of the NPRO capability. The manual told us that we should be able to scan the crystal temperature +/-16deg (about 30deg to 60deg).
What I like to try:
a) Change the NPRO temp to more cold side.
b) Change the MOPA head temp to a bit hot side.
c) Tweak the MOPA current (is it difficult?) |
2009
|
Sun Sep 27 15:25:58 2009 |
Koji | Update | PSL | SLOWscan result |
Oh, AMPMON dependence could be an artifact of the ND filter???
For my case, it should be real dependence on the NPRO wavelength,
as the other PDs like the PMC reflection (PMC_RFPDDC) and the RC reflection (FSS_RFPDDC) show the same dependence. |
2010
|
Sun Sep 27 23:21:14 2009 |
rana | Update | PSL | SLOWscan result |
Quote: |
What I like to try:
a) Change the NPRO temp to more cold side.
b) Change the MOPA head temp to a bit hot side.
c) Tweak the MOPA current (is it difficult?)
|
I think that the AMPMON ND problem was just that the responsivity changes with angle. So when I aligned it a little we got some few% improvement in the signal which is not a real power increase.
I don't think we can adjust any of the MOPA parameters because the controller is broken, but we can try the NPRO crystal temperature. |
2013
|
Mon Sep 28 17:39:34 2009 |
rob | Update | PSL | problems |
The PSL/IOO combo has not been behaving responsibly recently.
The first attachment is a 15 day trend of the MZ REFL, ISS INMON, and MC REFL power. These show two separate problems--recurring MZ flakiness, which may actually be a loose cable somewhere which makes the servo disengage. Such disengagement is not as obvious with the MZ as it is with other systems, because the MZ is relatively stable on its own. The second problem is more recent, just starting in the last few days. The MC is drifting off the fringe, either in alignment, length, or both. This is unacceptable.
The second attachment is a two-day trend of the MC REFL power. Last night I carefully put the beam on the center of the MC-WFS quads. This appears to have lessened the problem, but it has not eliminated it.
It's probably worth trying to re-measure the MCWFS system to make sure the control matrix is not degenerate. |
2033
|
Thu Oct 1 10:23:00 2009 |
steve | Update | PSL | MC2 damping restored again |
The EQ did not change the input beam pointing. All back to normal, except MC2 wachdogs tripped again. |
2049
|
Mon Oct 5 09:31:05 2009 |
Alberto | Frogs | PSL | PSL laser accidentally turned off |
Alberto, Steve,
While I was moving a cart near by the PSL table I pushed the red emergency button that turns off the PSL laser. We had to unlock the button and then power cycle the laser driver to turn the laser back on.
I relocked MZ, FSS, PMC and I'm now waiting for the power to finish ramping up back to the previous value. |
2050
|
Mon Oct 5 10:41:31 2009 |
Koji | Update | PSL | PSL laser accidentally turned off |
The PSL output looks smaller than the incident. Try to FSS Slow actuator adj of -5.6 (nominal), instead of -3.5.
Quote: |
Alberto, Steve,
While I was moving a cart near by the PSL table I pushed the red emergency button that turns off the PSL laser. We had to unlock the button and then power cycle the laser driver to turn the laser back on.
I relocked MZ, FSS, PMC and I'm now waiting for the power to finish ramping up back to the previous value.
|
|
2051
|
Mon Oct 5 13:43:37 2009 |
Zach | Update | PSL | PSL table photos |
I have been commissioned to take pictures of the PSL table so that it can be diagrammed. I am starting now (1:42 pm, 10/5/09). |
2052
|
Mon Oct 5 14:18:41 2009 |
Zach | Update | PSL | PSL table photos |
Quote:
|
I have been commissioned to take pictures of the PSL table so that it can be diagrammed. I am starting now (1:42 pm, 10/5/09).
|
All done (for now). That wasn't so bad, was it? |
2055
|
Mon Oct 5 19:39:26 2009 |
Koji | Update | PSL | PSL laser accidentally turned off |
I set the FSS slow actuator adj to -5.6 at the lunch time. It gave a little help at that time. Now max of the MC Trans is comming back somehow. I hope the MC Trans level is as good as before, if the HEPA is slowed down.
Quote: |
The PSL output looks smaller than the incident. Try to FSS Slow actuator adj of -5.6 (nominal), instead of -3.5.
Quote: |
Alberto, Steve,
While I was moving a cart near by the PSL table I pushed the red emergency button that turns off the PSL laser. We had to unlock the button and then power cycle the laser driver to turn the laser back on.
I relocked MZ, FSS, PMC and I'm now waiting for the power to finish ramping up back to the previous value.
|
|
|
2057
|
Tue Oct 6 10:09:55 2009 |
Zach | Update | PSL | More pictures |
I'm back to terrorize the PSL table again. The pictures I took yesterday were rubbish--today I'm using a clamp that Steve was nice enough to loan me. I'm starting now, at 10:09 am. |
2083
|
Mon Oct 12 18:37:55 2009 |
Zach | Update | PSL | Inventory |
--Apologies for the late post--
I was at the PSL table taking an inventory of the components for a while after Koji, Steve, and Kiwamu were there. I set the HEPAs back to 20% when I left (assuming that they were turned up when the compartment was opened). |
2088
|
Mon Oct 12 22:15:15 2009 |
rana | Configuration | PSL | Stray beam blocking |
You can remove the RFAM measuring setup. Once we upgrade, we will no longer have a MZ or the related problems. |
2089
|
Tue Oct 13 10:31:11 2009 |
Zach | Update | PSL | one more time |
I am at the PSL table taking what is hopefully the last set of pictures for the diagram. Woohoo. |
2090
|
Tue Oct 13 10:50:58 2009 |
Zach | Update | PSL | one more time |
Quote: |
I am at the PSL table taking what is hopefully the last set of pictures for the diagram. Woohoo.
|
I'm out, HEPAs are back at 20%. |
2098
|
Thu Oct 15 12:35:09 2009 |
Zach | Update | PSL | inventory |
I'm at the PSL table taking inventory of the elements I don't have down yet. |
2099
|
Thu Oct 15 12:57:23 2009 |
Zach | Update | PSL | inventory |
Quote: |
I'm at the PSL table taking inventory of the elements I don't have down yet.
|
OK, I'm out--hopefully for good. HEPAs back at 20%. |
2142
|
Mon Oct 26 15:40:01 2009 |
steve | Update | PSL | laser power is down |
The laser power is down 5-6% |
2147
|
Mon Oct 26 23:14:08 2009 |
Koji | Update | PSL | laser power is down |
I adjusted the steerings to the PMC and gained 7%. Now the MC_TRANS 7.0 has been recovered.
Actually I need another 7% to get MC_TRANS 7.5.
But I couldn't find how I can recover 126MOPA-AMPMON to 2.8ish.
Quote: |
The laser power is down 5-6%
|
|
2151
|
Tue Oct 27 18:01:49 2009 |
rob | Update | PSL | hmmm |
A 30-day trend of the PCDRIVE from the FSS. |
2155
|
Wed Oct 28 09:12:18 2009 |
steve | Update | PSL | PMC power on the rise? |
The PMC power is seems to be on the rise, ( MOPA_AMPMON is dropping ?) but I do not think it is real. We have Santa Anna wind condition, when the relative humidity drops and ......
There is an other funky think. The room temp became rock solid. The PSL HEPAs running at 20% and IFO-room ACs are also in normal operational mode. |
2158
|
Thu Oct 29 13:48:32 2009 |
Koji | Update | PSL | NPRO LTMP lowered 9.5deg |
13:00 Found MC TRANS less than 7.
13:50 Go into the PSL table.
14:20 Work done. Now I am running SLOWscan script.
15:10 SLOWscan finished. It was not satisfactory. I go into the table again.
15:15 Running SLOWscan again.
16:00 SLOWscan done. Lock PMC. Adjust NPRO current so as to maximize PMC TRANS.
16:10 Lock RC, PMC, MZ, MC. Align PMC / MZ on the table. Align MC WFS beams on the QPDs.
16:30 Work done.
New FSS-SLOWDC nominal is -4.0
Now MC TRANS is 7.9. This is +12% increase. ENJOY!
HEPA is on at 90%. Light is off.
---------
NPRO TEMP trimmer adjustment
o PSL NPRO TEMP trimmer at the back of the laser head was turned 6.5 times in CW.
o It reduced NPRO crystal temp by 9.5deg. (43.5deg -> 34.0deg for FSS_SLOWDC -5.5)
To revert the previous setting, refer to the former measurement
c.f. http://nodus.ligo.caltech.edu:8080/40m/2008
NPRO Thermal scan
o 2 scans are performed.
o I selected the colder side of the second scan. i.e. SLOWDC=-4.0
NPRO Current adjustment
o Tweaked C1:PSL-126MOPA_126CURADJ while looking at PMC TRANS.
o CURADJ was changed from -2.25 to -1.9. This corresponds to change of C1:PSL-126MOPA_CURMON from 2.503A to 2.547A. |
2161
|
Thu Oct 29 20:21:14 2009 |
Koji | Update | PSL | NPRO LTMP lowered 9.5deg |
Here is the plots for the powers. MC TRANS is still rising.
What I noticed was that C1:PSL-FSS_PCDRIVE nolonger hit the yellow alert.
The mean reduced from 0.4 to 0.3. This is good, at least for now. |
2165
|
Fri Oct 30 10:52:56 2009 |
Jenne | Update | PSL | HEPAs |
Zach found the HEPA switch on the PSL table OFF. He turned them on. |
2169
|
Mon Nov 2 13:34:36 2009 |
kiwamu | Configuration | PSL | removed multiply resonant EOM |
I removed the multiply resonant EOM that has been set by a SURF student from PSL table.
I will use it for checking the resonant circuit. |
2236
|
Wed Nov 11 12:29:44 2009 |
Alberto | Frogs | PSL | MC Locked on the wrong mode? |
This morning, after Steve pointed out that the readout RFAMPD_DC was zero, I thought of realigning the beam on the photodiode. Maybe I touched the lens or the beam splitter that send the beam on the diode when I installed an other beam splitter to make the measurement of the calibration between two ThorLabs PDA255 photodiodes.
After aligning the beam on the RFAMPD, the voltage of the DC readout was lower than it used to be (C1:IOO-RFAMPD_DC ~ 0.4 now vs. 4 as it was on November 4th).
I maximized the DC readout but the problem seems to be that the beam spot is not a round TEM00. In particular the spot looks like that of a TEM10 mode.
Since we're looking at the MC transmitted beam, is it possible that the MC is locked on the wrong mode?
Check out the attached picture. |
2249
|
Thu Nov 12 10:45:02 2009 |
Alberto | Update | PSL | Abandoned Frequency Generator |
This morning I found a frequency generator connected to something on the PSL table sitting on the blue step next to the sliding doors.
Is anyone using it? Has it been forgotten there? If that's the case, can the interested person please take care of removing it? |
2251
|
Thu Nov 12 11:19:10 2009 |
Koji | Update | PSL | Abandoned Frequency Generator |
Last night there was an activity for a calibratuon work, which I helped. I can take care of the FG.
Quote: |
This morning I found a frequency generator connected to something on the PSL table sitting on the blue step next to the sliding doors.
Is anyone using it? Has it been forgotten there? If that's the case, can the interested person please take care of removing it?
|
|
2256
|
Thu Nov 12 16:13:05 2009 |
Alberto | Update | PSL | MC Trans Offset |
On Rana's suggestion I checked the MC transmission QPD (C1:IOO-MC_TRANS_SUM). I found that the readout is almost zero when the MC is unlocked.
I unlocked the Mode Cleaner turning off the LSC control and disabling the autolocker. The QPD reads 0.014. It seems that there is no offset.
I also checked with the IR card around the photodetector and I didn't see any stray beam. |
2258
|
Thu Nov 12 17:15:43 2009 |
Koji | Update | PSL | MC Trans Offset |
OK. I have been keeping my eyes on the MC transmission. In deed, the MC trans has been well kept at around 7.7 since the last PSL work.
Even it was over the 8 today!
Quote: |
On Rana's suggestion I checked the MC transmission QPD (C1:IOO-MC_TRANS_SUM). I found that the readout is almost zero when the MC is unlocked.
I unlocked the Mode Cleaner turning off the LSC control and disabling the autolocker. The QPD reads 0.014. It seems that there is no offset.
I also checked with the IR card around the photodetector and I didn't see any stray beam.
|
|
2260
|
Thu Nov 12 17:42:04 2009 |
Koji | Update | PSL | MC Trans Offset |
PC_DRIVE is also improving after the last PSL work!
Quote: |
OK. I have been keeping my eyes on the MC transmission. In deed, the MC trans has been well kept at around 7.7 since the last PSL work.
Even it was over the 8 today!
|
|
2311
|
Mon Nov 23 00:46:09 2009 |
rana, rob | Update | PSL | ISS RIN: Its too high by 10x |
This plot shows the RIN as measured by the ISS. Its ~2 x 10^-7, whereas its supposed to be more like 3 x 10^-8.
The ISS has DC coupled RIN channels (with a _F suffix) and AC coupled RIN channels (with a _FW suffix). By using a swept sine, Rob determined that the AC coupled channels have an AC coupling pole at ~80 Hz. The attached plot uses this and then has the overall gain adjusted to match with the _F channels below 10 Hz.
The _F channels can be converted directly into RIN by just dividing the spectra by the mean value of the time series. The dark offset of these channels is small and so this only introduces a ~5-10% calibration error.
Question #1: Why is the RIN so bad? According to the MEDM screen, the photocurrent on the MON/SENS PDs is 1.9/1.3 mA. That's sort of low, but should still allow us to get 5x10^-8 in RIN.
Question #2: Does it make an effect on the current DC Readout work? IF so, should we try to fix up the ISS in a temporary way? Since the in-loop and out-of-loop detectors are completely coherent, all of the noise is likely just unsuppressed noise from the laser. We are unable to increase the gain because of the high frequency noise from the NPRO.
Let's remember to replace this ISS with a new one that can drive an AOM. Need a volunteer to get us a new ISS.
|
2335
|
Wed Nov 25 16:13:27 2009 |
rana | Update | PSL | Measured MC length--FSS trend |
but the increase in both the RCtrans and the RCrefl is consistent with my theory that the power going to the RC has increased ; its not just an increase in the visibility.
We should scan the AOM/VCO to make sure the frequency is matched to the resonance to within 0.5 dB. |
2336
|
Wed Nov 25 16:44:52 2009 |
Koji | Update | PSL | Measured MC length--FSS trend |
I checked C1:PSL-FSS_VCODETPWR. The attached is the 4 months trend of the FSS RCTRANS / RFPDDC(=FSS REFL) / VCODETPWR / VCOMODLEVEL.
Although VCO modulation level setting was mostly constnt, VCODETPWR, which presumably represents the RF level, changes time by time.
It coincides with the recent reduction of the RCTRANS/RFPDDC. Actually, my touch restored the VCO to the previous more stable state.
One can see that this is not only a single occation, but it happened before too. (In the middle of Aug.)
This could be explained by the bad contact of some cable or connector.
Nevertheless we need more careful investigation:
1. Understand what VCODETPWR is exactly.
2. Investigate relationship between VCOMODLEVEL / VCODETPWR / AOM deflection efficiency / RCTRANSPD
3. Confirm the frequency matching between the VCO and AOM.
Quote: |
but the increase in both the RCtrans and the RCrefl is consistent with my theory that the power going to the RC has increased ; its not just an increase in the visibility.
We should scan the AOM/VCO to make sure the frequency is matched to the resonance to within 0.5 dB.
|
|
2343
|
Sat Nov 28 20:27:12 2009 |
Koji | Update | PSL | FSS oscillation: Total gain reduced |
I stopped by the 40m for some reason and found that the MC trans was 7.5.
This was caused by an oscillation of FSS, which seemed to be started by itself.
The oscillation stopped by reducing the FSS total gain to +9dB (from +11dB).
This is not a permanent fix (i.e. autolocker will restore the gain).
If it seems necessary to reduce the FSS gain always, we change the MC autolocker script. |
2379
|
Thu Dec 10 09:51:06 2009 |
rob | Update | PSL | RCPID settings not saved |
Koji, Jenne, Rob
We found that the RCPID servo "setpoint" was not in the relevant saverestore.req file, and so when c1psl got rebooted earlier this week, this setting was left at zero. Thus, the RC got a bit chilly over the last few days. This channel has been added.
Also, RCPID channels have been added (manually) to conlog_channels. |
2381
|
Thu Dec 10 09:56:32 2009 |
Koji | Update | PSL | RCPID settings not saved |
Note: The set point C1:PSL-FSS_RCPID_SETPOINT is 37.0 on C1PSL_FSS_RCPID.adl.
Now the temp is recovering with its full speed. At some point we have to restore the value of the FSS SLOW DC as the temp change drag it up.
Quote: |
Koji, Jenne, Rob
We found that the RCPID servo "setpoint" was not in the relevant saverestore.req file, and so when c1psl got rebooted earlier this week, this setting was left at zero. Thus, the RC got a bit chilly over the last few days. This channel has been added.
Also, RCPID channels have been added (manually) to conlog_channels.
|
|
2466
|
Wed Dec 30 09:57:53 2009 |
steve | Update | PSL | thirsty water chiller |
I added 600 cc of Arrowhead Distilled Water to the chiller.
60 days plot shows that about every ~ 10 days I have to add some.
Please check the water level yourself. |
2563
|
Tue Feb 2 22:39:12 2010 |
Jenne | Update | PSL | IFO isn't playing nice tonight |
[Jenne, Kiwamu]
It's been an iffy last few hours here at the 40m. Kiwamu, Koji and I were all sitting at our desks, and the computers / RFM network decided to crash. We brought all of the computers back, but now the RefCav and PMC don't want to lock. I'm a wee bit confused by this. Both Kiwamu and I have given it a shot, and we can each get the ref cav to sit and flash, but we can't catch it. Also, when I bring the PMC slider rail to rail, we see no change in the PMC refl camera. Since c1psl had been finicky coming back the first time, I tried soft rebooting, and then keying the crate again, but the symptoms remained the same. Also, I tried burt restoring to several different times in the last few days, to see if that helped. It didn't. I did notice that MC2 was unhappy, which was a result of the burtrestores setting the MCL filters as if the cavity were locked, so I manually ran mcdown. Also, the MC autolocker script had died, so Kiwamu brought it back to life.
Since we've spent an hour on trying to relock the PSL cavities (the descriptive word I'm going to suggest for us is persistent, not losers), we're giving up in favor of waiting for expert advice in the morning. I suppose there's something obvious that we're missing, but we haven't found it yet...... |
2564
|
Wed Feb 3 01:17:19 2010 |
Koji | Update | PSL | IFO isn't playing nice tonight |
I checked the situation from my home and the problem was solved.
The main problem was undefined state of the autolocker and the strange undefined switch states, being associated with the bootfest and burtrestore.
- MC UP/DOWN status shows it was up and down. So I ran scripts/MC/mcup and scripts/MC/mcdown . These cleared the MC autolocker status.
- I had a problem handling the FSS. After mcup/mcdown above, I randomly pushed the "enable/disable" buttons and others, and with some reason, it recovered the handling. Actually it acquired the lock autonomously. Kiwamu may have also been working on it at the same time???
- Then, I checked the PSL loop. I disconnected the loop by pushing the "test" button. The DC slider changes the PZT voltage only 0~+24V. This is totally strange and I started pushing the buttons randomly. As soon as I pushed the "BLANK"/"NORMAL" button, the PZT output got back under the control.
- Then I locked the PMC, MZ, and MC as usual.
Alberto: You must be careful as the modulations were restored.
Quote: |
[Jenne, Kiwamu]
It's been an iffy last few hours here at the 40m. Kiwamu, Koji and I were all sitting at our desks, and the computers / RFM network decided to crash. We brought all of the computers back, but now the RefCav and PMC don't want to lock. I'm a wee bit confused by this. Both Kiwamu and I have given it a shot, and we can each get the ref cav to sit and flash, but we can't catch it. Also, when I bring the PMC slider rail to rail, we see no change in the PMC refl camera. Since c1psl had been finicky coming back the first time, I tried soft rebooting, and then keying the crate again, but the symptoms remained the same. Also, I tried burt restoring to several different times in the last few days, to see if that helped. It didn't. I did notice that MC2 was unhappy, which was a result of the burtrestores setting the MCL filters as if the cavity were locked, so I manually ran mcdown. Also, the MC autolocker script had died, so Kiwamu brought it back to life.
Since we've spent an hour on trying to relock the PSL cavities (the descriptive word I'm going to suggest for us is persistent, not losers), we're giving up in favor of waiting for expert advice in the morning. I suppose there's something obvious that we're missing, but we haven't found it yet......
|
|
2565
|
Wed Feb 3 07:57:01 2010 |
steve | Update | PSL | PMC transmission is low |
The low PMC transmission alarm was on this morning. The PMC alignment needs a touch up. |
2650
|
Tue Mar 2 12:20:54 2010 |
kiwamu | Update | PSL | stray beam |
In order to block stray beams, I have put some beam dumps and razor blades on the PSL table.
There were three undesired spots in total. I found two spots on the south side door of the PSL room, close to Mach-Zehnder.
Another spots was on the middle of the north door. Now they all are blocked successfully. |
2690
|
Sun Mar 21 20:08:20 2010 |
kiwamu, rana | Update | PSL | EOM wasit size |
We are going to set the waist size to 0.1 mm for the beam going through the triple resonant EOM on a new PSL setup.
When we were drawing a new PSL diagram, we just needed to know the waist size at the EOM in order to think about mode matching.

This figure shows the relation between the waist size and the spot size at the aperture of the EOM.
The x-axis is the waist size, the y-axis is the spot size. It is clear that there is a big clearance at 0.1 mm waist size. This is good.
Also it is good because the waist size is much above the damage threshold of the EO crystal (assuming 1W input).
The attached file is the python code for making this plot. |
2691
|
Sun Mar 21 21:02:39 2010 |
Koji | Update | PSL | EOM waist size |
You don't need a lengthy code for this. It is obvious that the spot size at the distance L is minimum when L =
zR, where zR is the Rayleigh range. That's all.
Then compare the spot size and the aperture size whether it is enough or not.
It is not your case, but if the damage is the matter, just escape to the large zR side. If that is not possible
because of the aperture size, your EOM is not adequate for your purpose. |
2720
|
Sun Mar 28 20:05:33 2010 |
rana | Summary | PSL | FSS Work from Sunday: AOM/VCO level set wrong |
Just before working on the FSS today, I noticed that the VCO RF output level was set incorrectly.
This should ALWAYS be set so as to give the maximum power in the first order diffracted sideband. One should set this by maximizing the out of lock FSS RFPD DC level to max.
The value was at 2.8 on the VCOMODLEVEL slider. In the attached plot (taken with the FSS input disabled) you can see that this puts us in the regime where the output power to the FSS is first order sensitive to the amplitude noise on the electrical signal to the AOM. This is an untenable situation.
For adjusting the power level to the FSS, we must always use the lamba/2 plate between the AOM and the RC steering mirrors. This dumps power out to the side via a PBS just before the periscope. |
2721
|
Sun Mar 28 20:51:31 2010 |
rana | Summary | PSL | FSS Work from Sunday: Cavity Suspension is Ridiculously Undamped! |
What is the Transfer Function of the suspension of the reference cavity? What were the design requirements? What is the Q and how well does the eddy current damping work? What did Wolfowitz know about the WMD and when? Who cooked the RTV in there and why didn't we use Viton??
To get to the bottom of these questions, today I shook the cavity and measured the response. To read out the pitch and yaw modes separately, I aligned the input beam to be misaligned to the cavity. If the beam is mis-aligned in yaw, for example, the transmitted light power becomes first order sensitive to the yaw motion of the cavity.
In the attached image (10 minute second-trend), you can see the second trends for the transmitted and relfected power. The first ringdown comes from the pitch or vertical mode. The second (shorter) one comes from the yaw misalignment and the yaw shake.
To achieve the up/down shake, I leaned onto the table and pumped it at its eigenfrequency. For the yaw shake, I put two fingers on the RC can's sweater and pushed with several pounds of force at the yaw eigenfrequency (2.6 Hz). For the vertical, I jumped up and down at half the vertical eigenfrequency (4 Hz).
I also made sure that the .SCAN field on these EPICS records were set to 9 so that there is no serious effect from a beating between the eigenfrequency and the EPICS sample rate.
Punchline:
f_vert = 4 Hz
tau_vert = 90 seconds
Q_vert = 1000 (yes, that number over there has 3 zeros)
f_hor = 2.6 Hz
tau_hor = 30 seconds
Q_hor = 250
This is an absurd and probably makes us very sensitive to seismic noise - let's make sure to open up the can and put some real rubber in there to damp it. My guess is that these high Q modes