  6944   Mon Jul 9 11:27:27 2012 JenneUpdateComputersc1oaf has been down for several days - BURT restore wasn't done correctly on startup

The c1oaf model hasn't been running for a few days (since the leap second problems we were having last week).  I had looked into it, but finally figured it out (with Jamie's help) today. 

The BURT restore has to be given to the model during startup, but for whatever reason it wasn't BURT restoring until *after* the model had already failed to start.  The symptoms were:  no 'heartbeat' for the oaf model, no connection to the fb, NO SYNC on the GDS screen, 0x4000.  the BURT restore button was green, which threw me off the scent, but that's just because it did, in fact, get set, just way too late.

I ended up looking in the dmesg of the lsc computer, and the last set of stuff was several lines of "[3354303.626446] c1oaf: Epics burt restore is 0".  Nothing else was written after that.  Jamie pointed out that this meant the BURT restore wasn't getting sent before the model unloaded itself and decided not to run.

The solution:  restart the model, and manually click the BURT restore button as soon as you're able (after everything comes back from being white).  We used to have to do this, but then there was a "fix", which apparently isn't super robust and failed for the oaf (even though it used to work just fine).  Bugzilla report submitted.

  9911   Mon May 5 19:51:56 2014 jamieUpdateCDSc1oaf model broken because of broken BLRMS block

I finally tracked down the problem with the c1oaf model to the BLRMS part:

/opt/rtcds/userapps/release/cds/common/models/BLRMS.mdl

blrms-hot-mess.pngsddefault.jpg

Note that this is pulling from a cds/common location, so presumably this is a part that's also being used at the sites.

Either there was an svn up that pulled in something new and broken, or the local version is broken, or who knows what.

We'll have to figure how what's going on here, but in the mean time, as I already mentioned, I'm leaving the c1oaf model off for now.

 RXA: also...we updated Ottavia to Ubuntu 12 LTS...but now it has no working network connection. Needs help.  (which of course has nothing whatsoever to do with this point )

  14922   Wed Oct 2 10:40:07 2019 gautamUpdateCDSc1oaf model restarted

This morning, I restarted the c1oaf model on the c1lsc machine, so as to have the option of enabling some feedforward action. Unsurprisingly, the "DC" indicator is red, citing a "0x2bad". In the past, I've been able to correct this by simply restarting the model. But given the fragility of the c1lsc machine, I think I'll live with not having the OAF model signals in frames. Medium-term, I'd like to pare down the c1oaf model a bit - I think it has way too many options/matrices right now, and is an un-necessarily bloated and heavy model. Unless there are serious objections, I will do this work when I next feel like it.

  14522   Mon Apr 8 11:53:17 2019 gautamUpdateCDSc1oaf needs debugging

I tried restarting c1oaf this weekend to see if turning on the MC length FF would affect the ALS noise performance. I burtrestored the filter settings from March 2016. However, I noticed several possible anomalies, which need debugging. I am not turning the model off because of the possibility of having to reboot all the vertex FEs, but this model is totally unusable right now.

  1. Attachment #1 - the vertex seismometer input produces 1e+20 cts at the output of the feedforward filter. Attachment #2 shows the shape of the feedforward filters - doesn't explain the saturation. Since this is a feedforward loop, a runaway loop can't be the explanation either.
  2. The MC length feedforward control signal is supposed to only go to MC2 - but MC1 and MC3 coil outputs were saturated when I enabled the feedforward.
  7287   Mon Aug 27 17:14:00 2012 jamieUpdateCDSc1oaf problem

Quote:

I came in to the lab in the evening and found c1lsc had "red" for FB connection.
I restarted c1lsc models and it kept hung the machine everytime.

I decided to kill all of the model during the startup sequence right after the reboot.
Then run only c1x04 and c1lsc. It seems that c1oaf was the cause, but it wasn't clear.

The "red for FB connection" issue was probably a dead mx_stream on c1lsc.  That can usually be fixed by just restarting mx_stream.

There is definitely a problem with c1oaf, though.  It crashes immediately after attempting to start.  kernel log for a crash included below.

We will leave c1oaf off until we have time to debug.

[83752.505720] c1oaf: Send Computer Number  = 0
[83752.505720] c1oaf: entering the loop
[83752.505720] c1oaf: waiting to sync 19520
[83753.207372] c1oaf: Synched 701492
[83753.207372] general protection fault: 0000 [#2] SMP 
[83753.207372] last sysfs file: /sys/devices/pci0000:00/0000:00:1e.0/0000:2e:01.0/class
[83753.207372] CPU 4 
[83753.207372] Modules linked in: c1oaf c1ass c1sup c1lsp c1cal c1lsc c1x04 open_mx dis_irm dis_dx dis_kosif mbuf [last unloaded: c1oaf]
[83753.207372] 
[83753.207372] Pid: 0, comm: swapper Tainted: G      D    2.6.34.1 #5 X7DWU/X7DWU
[83753.207372] RIP: 0010:[<ffffffffa1bf7567>]  [<ffffffffa1bf7567>] T.2870+0x27/0xbf0 [c1oaf]
[83753.207372] RSP: 0000:ffff88023ecc1aa8  EFLAGS: 00010092
[83753.207372] RAX: ffff88023ecc1af8 RBX: ffff88023ecc1ae8 RCX: ffffffffa1c35e48
[83753.207372] RDX: 0000000000000000 RSI: 0000000000000020 RDI: ffffffffa1c21360
[83753.207372] RBP: ffff88023ecc1bb8 R08: 0000000000000000 R09: 0000000000175f60
[83753.207372] R10: 0000000000000000 R11: ffffffffa1c2a640 R12: ffff88023ecc1b38
[83753.207372] R13: ffffffffa1c2a640 R14: 0000000000007fff R15: 0000000000000000
[83753.207372] FS:  0000000000000000(0000) GS:ffff880001f00000(0000) knlGS:0000000000000000
[83753.207372] CS:  0010 DS: 0000 ES: 0000 CR0: 0000000080050033
[83753.207372] CR2: 000000000378a040 CR3: 0000000001a09000 CR4: 00000000000406e0
[83753.207372] DR0: 0000000000000000 DR1: 0000000000000000 DR2: 0000000000000000
[83753.207372] DR3: 0000000000000000 DR6: 00000000ffff0ff0 DR7: 0000000000000400
[83753.207372] Process swapper (pid: 0, threadinfo ffff88023ecc0000, task ffff88023ec7eae0)
[83753.207372] Stack:
[83753.207372]  ffff88023ecc1ab8 0000000000000096 0000000000000019 ffff88023ecc1b18
[83753.207372] <0> 0000000000014729 0000000000032a0c ffff880001e12d90 000000000000000a
[83753.207372] <0> ffff88023ecc1bb8 ffffffffa1c06cad ffff88023ecc1be8 000000000000000f
[83753.207372] Call Trace:
[83753.207372]  [<ffffffffa1c06cad>] ? filterModuleD+0xd6d/0xe40 [c1oaf]
[83753.207372]  [<ffffffffa1c07ae3>] feCode+0xd63/0x129b0 [c1oaf]
[83753.207372]  [<ffffffffa1c00dc6>] ? T.2888+0x1966/0x1f10 [c1oaf]
[83753.207372]  [<ffffffffa1c1b3bf>] fe_start+0x1c8f/0x3060 [c1oaf]
[83753.207372]  [<ffffffff8102ce57>] ? select_task_rq_fair+0x2c8/0x821
[83753.207372]  [<ffffffff8104cd8b>] ? enqueue_hrtimer+0x65/0x72
[83753.207372]  [<ffffffff8104d8f6>] ? __hrtimer_start_range_ns+0x2d6/0x2e8
[83753.207372]  [<ffffffff8104d91b>] ? hrtimer_start+0x13/0x15
[83753.207372]  [<ffffffff810173df>] play_dead_common+0x6e/0x70
[83753.207372]  [<ffffffff810173ea>] native_play_dead+0x9/0x20
[83753.207372]  [<ffffffff81001c38>] cpu_idle+0x46/0x8d
[83753.207372]  [<ffffffff814ec523>] start_secondary+0x192/0x196
[83753.207372] Code: 1f 44 00 00 55 66 0f 57 c0 48 89 e5 41 57 41 56 41 55 41 54 53 48 8d 9d 30 ff ff ff 48 8d 43 10 4c 8d 63 50 48 81 ec e8 00 00 00 <66> 0f 29 85 30 ff ff ff 48 89 85 18 ff ff ff 31 c0 48 8d 53 78 
[83753.207372] RIP  [<ffffffffa1bf7567>] T.2870+0x27/0xbf0 [c1oaf]
[83753.207372]  RSP <ffff88023ecc1aa8>
[83753.207372] ---[ end trace df3ef089d7e64971 ]---
[83753.207372] Kernel panic - not syncing: Attempted to kill the idle task!
[83753.207372] Pid: 0, comm: swapper Tainted: G      D    2.6.34.1 #5
[83753.207372] Call Trace:
[83753.207372]  [<ffffffff814ef6f4>] panic+0x73/0xe8
[83753.207372]  [<ffffffff81063c19>] ? crash_kexec+0xef/0xf9
[83753.207372]  [<ffffffff8103a386>] do_exit+0x6d/0x712
[83753.207372]  [<ffffffff81037311>] ? spin_unlock_irqrestore+0x9/0xb
[83753.207372]  [<ffffffff81037f1b>] ? kmsg_dump+0x115/0x12f
[83753.207372]  [<ffffffff81006583>] oops_end+0xb1/0xb9
[83753.207372]  [<ffffffff8100674e>] die+0x55/0x5e
[83753.207372]  [<ffffffff81004496>] do_general_protection+0x12a/0x132
[83753.207372]  [<ffffffff814f17af>] general_protection+0x1f/0x30
[83753.207372]  [<ffffffffa1bf7567>] ? T.2870+0x27/0xbf0 [c1oaf]
[83753.207372]  [<ffffffffa1c06cad>] ? filterModuleD+0xd6d/0xe40 [c1oaf]
[83753.207372]  [<ffffffffa1c07ae3>] feCode+0xd63/0x129b0 [c1oaf]
[83753.207372]  [<ffffffffa1c00dc6>] ? T.2888+0x1966/0x1f10 [c1oaf]
[83753.207372]  [<ffffffffa1c1b3bf>] fe_start+0x1c8f/0x3060 [c1oaf]
[83753.207372]  [<ffffffff8102ce57>] ? select_task_rq_fair+0x2c8/0x821
[83753.207372]  [<ffffffff8104cd8b>] ? enqueue_hrtimer+0x65/0x72
[83753.207372]  [<ffffffff8104d8f6>] ? __hrtimer_start_range_ns+0x2d6/0x2e8
[83753.207372]  [<ffffffff8104d91b>] ? hrtimer_start+0x13/0x15
[83753.207372]  [<ffffffff810173df>] play_dead_common+0x6e/0x70
[83753.207372]  [<ffffffff810173ea>] native_play_dead+0x9/0x20
[83753.207372]  [<ffffffff81001c38>] cpu_idle+0x46/0x8d
[83753.207372]  [<ffffffff814ec523>] start_secondary+0x192/0x196

  11551   Tue Sep 1 02:44:44 2015 KojiSummaryCDSc1oaf, c1mcs modified for the IMC angular FF

[Koji, Ignacio]

In order to allow us to work on the IMC angular FF, we made the signal paths from PEM to MC SUSs.
In fact, there already were the paths from c1pem to c1oaf. So, the new paths were made from c1oaf to c1mcs. (Attachment 1~3)

After some debugging those two models started running. The additional cost of the processing time is insignificant.
FB was restarted to accomodate the change.

Once the modification of the models was completed, the OAF screens were modified. It seemed that the Kissel button
for the output matrix haven't been updated for the PRM ASC implementation. This was fixed as the button was updated this time.
In addition, the button for the FM matrix was also made and pasted.

 

  14123   Wed Aug 1 20:44:57 2018 gautamSummaryComputersc1omc model (re?)created

The main motivation behind adding a DAC card in c1ioo was to setup an RTCDS model for the OMC. Attachment #1 shows the new look CDS overview screen. Here is what I did.

Mostly, I followed instructions from when I setup the model for the EX green PZTs.


Simulink model:

The model is just a toy for now (CDS parameters, ADC block and 2 CDS filter modules). I leave it to Aaron to actually populate it, check functionality etc. The path to the model is /opt/rtcds/caltech/c1/userapps/release/isc/c1/models/c1omc.mdl. I am listing the parameters set on the CDS_PARAMETERS block:

  • host = c1ioo
  • site = c1
  • rate = 16k
  • dcuid = 27 (which I chose after making sure that this dcuid was not used on this list which I also updated by adding c1omc and moving c1imc to "old")
  • specific_cpu = 6 (again chosen after checking the available CPUs in the above list and confirming using the cset utility).
  • adc_Slave = 1
  • shmem_daq = 1
  • no_rfm_dma = 1
  • biquad = 1

Building and installing model:

Once the model was installed, I logged into c1ioo, and built and installed the models using the usual rtcds make and rtcds install instructions. Before starting the model, I edited /diskless/root.jessie/etc/rtsystab to allow c1omc to be run on c1ioo. Using sudo cset set, I verified that CPU #6 is no longer listed (if I understand correctly, the RTCDS system takes over the core).


MEDM:

To reflect all this on the MEDM CDS OVERVIEW screen, I just edited the screen.

  • Moved the orange explanation of bits over to the c1iscey panel to make space in the c1ioo panel.
  • Edited the macros to reflect the c1omc parameters.

DAQD:

Finally, I followed the instructions here to get the channels into frames and make all the indicators green. Went into fb and restarted the daqd processes. All looks good smiley. I'm going to leave the model running overnight to investigate stability. I forgot to svn commit the model tonight, will do it tomorrow.


The testing plan (at least initially) is to install the AA and AI boards from the OMC rack in 1X1/1X2. Then we will have short SCSI cables running from the ADC/DAC to these. The actual HV driving stages will remain in the OMC rack (NE corner of AS table).

@Steve, can we get 10 Male-Female D9 cables so that we can run them from 1X1/1X2 to the OMC rack?


Unrelated to this work: There were 2 crashes of the models on c1lsc, one ~6pm and one right now ~1030pm. The restart script brought everything back gracefully  yes...

  14126   Thu Aug 2 20:54:18 2018 gautamSummaryComputersc1omc model looks stable

Actually, c1lsc had crashed again sometime last night so I had to reboot everything this morning. I used the reboot script again, but I increased the sleep time between trying to start up the models again so that I could walk into the VEA and power cycle the c1lsc expansion chassis, as this kind of frequent model crash has been fixed by doing so in the past. Sure enough, there have been no issues since I rebooted everything at ~1030 in the morning. 

The c1omc model itself has been stable as well, though of course, there is nothing in there at the moment. I may do a check of the newly installed DAC tomorrow just to see that we can put out a sine wave.

Steve has ordered the D-sub cabling that will allow us to route signals between AA/AI boards in 1X1/1X2 to the HV PZT electronics in the OMC rack. Things look setup for a measurement next week. Aaron will post a block diagram + photoz of what box goes where in the electronics racks.

  2348   Mon Nov 30 16:23:51 2009 JenneUpdateComputersc1omc restarted

I found the FEsync light on the OMC GDS screen red.  I power cycled C1OMC, and restarted the front end code and the tpman.  I assume this is a remnant of the bootfest of the morning/weekend, and the omc just got forgotten earlier today.

  5971   Mon Nov 21 17:07:34 2011 MirkoUpdateCDSc1pem model dead

For some reason C1PEM doesn't seem to work anymore after a recompilation. It did recompile fine. We just changed some channel / subsystem names.

Tried reverting to the svn version. Doesn't work. Reboot C1SUS also no good.

  5973   Mon Nov 21 22:51:55 2011 MirkoUpdateCDSc1pem model dead

Quote:

For some reason C1PEM doesn't seem to work anymore after a recompilation. It did recompile fine. We just changed some channel / subsystem names.

Tried reverting to the svn version. Doesn't work. Reboot C1SUS also no good.

 It is fine again. Thanks Jamie.

  4028   Wed Dec 8 14:51:09 2010 josephbUpdateCDSc1pem now recording data

Problem:

c1pem model was reporting all zeros for all the PEM channels.

Solution:

Two fold.  On the software end, I added ADCs 0, 1, and 2 to the model.  ADC 3 was already present and is the actual ADC taking in PEM information.

There was a problem noted awhile back by Alex and Rolf that there's a problem with the way the DACs and ADCs are number internally in the code.  Missing ADCs or DACs prior to the one you're actually using can cause problems.

At some point that problem should be fixed by the CDS crew, but for now, always include all ADCs and DACs up to and including the highest number ADC/DAC you need to use for that model.

On the physical end, I checked the AA filter chassis and found the power was not plugged in.  I plugged it in.

Status:

We now have PEM channels being recorded by the FB, which should make Jenne happier.

  4101   Sat Jan 1 19:13:40 2011 ranaUpdateCDSc1pem now recording data

 I found that there was no PEM data nor any other data (no SUS or otherwise. No testpoints, no DAQ).

I went through the procedure that Jenne has detailed in the Wiki but it didn't work.

1) Firstly, the 'telnet fb 8088' step doesn't work. It says "Connected to fb.martian" but then just hangs. To replicate the effect of this step I tried ssh'ing to fb and doing a 'pkill daqd'. That works to restart the daqd process.

2) The wiki instructions had a problem. In the GUI step, it should say 'Save' after the Acquire bit has been set to 1. Even so, this works to get the .ini file right and the DTT can see the correct channel list, but none of the channels are available. There are just 'Unable to obtain measurement data'.

3) I tried running 'startc1pem', but no luck. I also tried rebooting c1sus from the command line. That worked so far as to come back up with all the right processes running, but still no data. The actual /frames directory shows that there are frames, but we just can't see the data. I also tried to get data usind the DTT-NDS2 method, but still no luck. (*** ITMX and ITMY both came back with all their filters off; worth checking if their BURTs are working correctly.)

Using DataViewer, however, I AM able to see the data (although the channel name is RED). In fact, I am able to see the trend data ever since I changed the Acquire bit to 1. Plot attached as evidence. Why does DTT not work anymore???

  12592   Wed Nov 2 22:56:45 2016 gautamUpdateCDSc1pem revamped

installing the BLRMS 2k blocks turned out to be quite non-trivial due to a whole host of CDS issues that had to be debugged, but i've restored everything to a good state now, and the channels are being logged. detailed entry with all the changes to follow.

  12595   Thu Nov 3 12:38:42 2016 gautamUpdateCDSc1pem revamped

A number of changes were made to C1PEM and some library parts. Recall that the motivation was to add BLRMS channels for all our suspension coils and shadow sensor PDs, which we are first testing out on the IMC mirrors.

Here is the summary:

BLRMS_2k library block

  • The name of the custom C code block in this library part was named 'BLRMSFILTER' which conflicted with the name of the function call in the C code it is linked to, which lead to compilation errors
  • Even though the part was found in /opt/rtcds/userapps/release/cds/c1/models and not in the common repository, just to be safe, I made a copy of the part called BLRMS_2k_40m which lives in the above directory. I also made a copy of the code it calls in /opt/rtcds/userapps/release/cds/c1/src

C1PEM model + filter channels

  • Adding the updated BLRMS_2k_40m library part still resulted in some compilation errors - specifically, it was telling me to check for missing links around the ADC parts
  • Eric suggested that the error messages might not be faithfully reporting what the problem is - true enough, the problem lay in the fact that c1pem wasn't updated to follow the namespace convention that we now use in all the RT models - the compiler was getting confused by the fact that the BLRMS stuff was in a namespace block called 'SUS', but the rest of the PEM stuff wasn't in such a block
  • I revamped c1pem to add namespace blocks called PEM and DAF, and put the appropriate stuff in the blocks, after which there were no more compilation errors
  • However, this namespace convention messed up the names of the filter modules and associated channels - this was resolved with Eric's help (find and replace did the job, this is a familiar problem that we had encountered not too long ago when C1IOO was similarly revamped...)
  • There was one last twist in that the model would compile and install, but just would not start. I tried the usual voodo of restarting all the models, and even did a soft reboot of c1sus, to no avail. Looking at dmesg, I tracked the problem down to a burt restore issue - the solution was to press the little 'BURT' button next to c1pem on the CDS overview MEDM screen as soon as it appeared while restarting the model

All the channels seem to exist, and FB seems to not be overloaded judging by the performance overnight up till the power outage. I will continue to monitor this...

GV Edit 3 Nov 2016 7pm:

I had meant to check the suitability of the filters used - there is a detailed account of the filters implemented in BLRMSFILTER.c here, and I quickly looked at the file on hand to make sure the BP filters made sense (see Attachment #1). These the BP filters are 8th order elliptical filters and the lowpass filters are16th order elliptical filters scaled for the appropriate frequency band, which are somewhat different from what we use on the seismometer BLRMS channels, where the filters are order 4, but I don't think we are significantly overloaded on the computational aspect, and the lowpass filters have sufficiently steep roll-off, these should be okay...

  12597   Thu Nov 3 13:36:16 2016 ericqUpdateCDSc1pem revamped

It seems that the EX and EY BLRMS banks were missing the BP and LP filters for the 0.03-0.1 and 0.1-0.3 bands. I've copied over the filters from the BS seismometer.

However, if it looks like the integrated C code BLRMS block works out well, we could replace the seismometers' filter module heavy BLRMS blocks and cut down on the PEM model bloat.

  12766   Fri Jan 27 21:21:35 2017 gautamUpdateCDSc1pem revamped

The coil and PD BLRMS are useful tools in identifying when glitches occur in the PD  readout, I thought it would be good to install them for ITMY, ETMX and SRM (since I plan to switch the MC3 satellite box, which we suspect to be problematic, with the SRM one). For this purpose, I had to install some IPC SHMEM blocks in C1SUS and recompile. 24 IPC channels were added to pipe the coil, PD and Oplev signals from C1SUS to C1PEM - the recompilation went smoothly, and it doesn't look like the model computation time has increased significantly or that the model is any closer to timing out.

However, I was unable to install the BLRMS blocks in C1PEM, as when I tried to compile the model with BLRMS for these extra 24 channels, I got a compilation error saying that I have exceeded the maximum allowed 499 testpoints per channel. Is there any workaround to this? It would be possible to create a custom BLRMS block that doesn't have all those testpoints, maybe this is the way to go? Especially if we want to install these channels for all our SOS optics, and also replace the current Seismic BLRMS with this scheme for consistency?

GV edit: I have implemented this scheme - after backing up the original BLRMS_2k part, I made a new one with no testpoints and only EPICS readouts. Doing so allowed me to recompile c1pem without any issues, the CPU time seems to have gone up by 3us from ~55us to 58us. So the BLRMS data record is only available at 16Hz, since there are no DQ channels in the BRLMS block - do we want these in any case? Let's see how this does over the weekend...

  11882   Mon Dec 14 23:56:29 2015 ericqUpdateCDSc1pem reverted

To get C1PEM data back into the frames, I removed the new BLRMS blocks, recompiled, reinstalled, re-enabled it in daqd, restarted.

We still really want more headroom in our framebuilder situation. 

  5549   Mon Sep 26 17:49:51 2011 KojiUpdatePSLc1psl

[Koji Suresh]

c1psl has got frozen during our ezcaread/write business.
After the target was rebooted and we lost the previous setting as there was no burt snapshot for the slow targets since Dec 13, 2010.

It seems that burtrestore is essential for the bootstrapping of the MC servo, as the auto locker script refers the locking parameters
from the PSL setting values (C1PSL_SETTINGS_SET.adl).

Jenne is working on the recovery of the snap-shotting for the slow targets.

  15238   Mon Mar 2 16:29:40 2020 gautamUpdateElectronicsc1psl VME crate removed, Acro-crate installed

[JV, JWR, YD, GV]

  • The old c1psl VME crate, and all the ribbon cables connected to it were removed from 1X1. They are presently dumped in the office area - we will clear these in the next few days, once the c1iool0 crate also gets removed from the rack.
  • The Acromag crate was capped on the top and bottom, had ears bolted on, and was installed on support rails in the newly cleared up space.
  • The strange orientation of the crate (with the intended backside facing the front of the rack) is to facilitate easy access to the "spare" channels we have in this box, e.g. for a future ISS or laser amplifier.
  • Remaining connections to make are (these will be done tomorrow along with the extrication of the c1iool0 VME crate):
    • PMC trans PD
    • FSS RMTEMP 
    • PSL shutter
    • 2W Mephisto diagnostic connector
    • 24 V DC from Sorensens via DIN connector (we are waiting on a new power cable to arrive).
  15194   Thu Feb 6 21:54:13 2020 JonUpdatePSLc1psl bench testing complete

Today I engineered the last piece of the new c1psl system: the multi-bit binary output (mbbo) channels that control the MC servo board gains. These 6-bit channels have to be split across two 4-bit Acromag registers. To enforce synchronous switching, I adapted the latch.py script developed by Gautam to address this problem in c1iscaux. Analogously to the c1iscaux implementation, I scripted the code to automatically run as a systemd service which is launched by the main modbusIOC service. I tested this all using the DB37 LED test board and confirmed it to work.

This now completes the electronics bench testing.

There are still several DB37 connectors to be wired, which carry only spare channels for future use and are not interfaced with the EPICS IOC. Jordan and I discussed this today and he or Chub will complete it shortly. To allow time for the spare channel wiring to be completed (as well as for more locking progress before interruption), Gautam and I think Monday/Tuesday next week would be the earliest possible window to install the new system.

  12852   Fri Feb 24 20:38:01 2017 johannesUpdateComputersc1psl boot-stall culprit identified

[Gautam, Johannes]

c1psl finally booted up again, PMC and IMC are locked.

Trying to identify the hickup from the source code was fruitless. However, since the PMCTRANSPD channel acqusition failure occured long before the actual slow machine crashed, and since the hickup in the boot seemed to indicate a problem with daughter module identification, we started removing the DIO and DAQ modules:

  1. Started with the ones whose fail LED stayed lit during the boot process: the DIN (XVME-212) and the three DACs (VMIVME4113). No change.
  2. Also removed the DOUT (XVME-220) and the two ADCs (VMIVME 3113A and VMIVME3123). It boots just fine and can be telnetted into!
  3. Pushed the DIN and the DACs back in. Still boots.
  4. Pushed only VMIVME3123 back in. Boot stalls again.
  5. Removed VMIVME3123, pushed VMIVME 3113A back in. Boots successfully.
  6. Left VMIVME3123 loose in the crate without electrical contact for now.
  7. Proceeded to lock PMC and IMC

The particle counter channel should be working again.

  • VMIVME3123 is a 16-Bit High-Throughput Analog Input Board, 16 Channels with Simultaneous Sample-and-Hold Inputs
  • VMIVME3113A is a Scanning 12-Bit Analog-to-Digital Converter Module with 64 channels

/cvs/cds/caltech/target/c1psl/psl.db lists the following channels for VMIVME3123:

Channels currently in use (and therefore not available in the medm screens):

  • C1:PSL-FSS_SLOW_MON
  • C1:PSL-PMC_PMCERR
  • C1:PSL-FSS_SLOWM
  • C1:PSL-FSS_MIXERM
  • C1:PSL-FSS_RMTEMP
  • C1:PSL-PMC_PMCTRANSPD

Channels not currently in use (?):

  • C1:PSL-FSS_MINCOMEAS
  • C1:PSL-FSS_RCTRANSPD
  • C1:PSL-126MOPA_126MON
  • C1:PSL-126MOPA_AMPMON
  • C1:PSL-FSS_TIDALINPUT
  • C1:PSL-FSS_TIDALSET
  • C1:PSL-FSS_RCTEMP
  • C1:PSL-PPKTP_TEMP

There are plenty of channels available on the asynchronous ADC, so we could wire the relevant ones there if we done care about the 16 bit synchronous sampling (required for proper functionality?)

Alternatively, we could prioritize the Acromag upgrade on c1psl (DAQ would still be asynchronous, though). The PCBs are coming in next Monday and the front panels on Tuesday.

 

 

Some more info that might come in handy to someone someday:

The (nameless?) Windows 7 laptop that lives near MC2 and is used for the USB microscope was used for interfacing with c1psl. No special drivers were necessary to use the USB to RS232 adapter, and the RJ45 end of the grey homemade DB9 to RJ45 cable was plugged into the top port which is labeled "console 1". I downloaded the program "CoolTerm" from http://freeware.the-meiers.org/#CoolTerm, which is a serial protocol emulator, and it worked out of the box with the adapter. The standard settings fine worked for communicating with c1psl, only a small modification was necessary: in Options>Terminal make sure that "Enter Key Emulation" is set from "CR+LF" to "CR", otherwise each time 'Enter' is pressed it is actually sent twice.

  15150   Thu Jan 23 23:07:04 2020 JonConfigurationPSLc1psl breakout board wiring

To facilitate wiring the c1psl chassis and scripting loopback tests, I've compiled a distilled spreadsheet with the Acromag-to-breakout board wiring, broken down by connector. This information is extractable from the master spreadsheet, but not easily. There were also a few apparent typos which are fixed here.

The wiring assignments at the time of writing are attached below. Here is the link to the latest spreadsheet.

  15117   Mon Jan 13 15:47:37 2020 shrutiConfigurationComputer Scripts / Programsc1psl burt restore

[Yehonathan, Jon, Shruti]

Since the PMC would not lock, we initially burt-restored the c1psl machine to the last available shapshot (Dec 10th 2019), but it still would not lock.

Then, it was burt-restored to midnight of Dec 1st, 2019, after which it could be locked.

  13742   Mon Apr 9 23:28:49 2018 johannesConfigurationDAQc1psl channel list

I made a list of all the physical c1psl channels to get a better idea for how many acromags we need to replace it eventually. There  3123 unit is the one whose failure had prevented c1psl from booting, which is why it was unplugged (elog post 12852), and its channels have been inactive since. Are the 126MOPA channels used for the current mephisto? 126 tells me it's for an old lightwave laser, but I was checking a few and found that they have non-zero, changing values, so they may have been rewired.

It also hosts some virtual channels for the ISS with root C1:PSL-ISS_ defined in iss.db and dc.db, the PSL particle counter with root C1:PEM- defined in PCount.db  and a whole lot of PSL status channels defined in pslstatus.db. Transfering these virtual channels to a different machine is almost trivial, but the serial readout of the particle counter would have to find a new home.

Long story short - we need:

Function Type # Channels #Channels (no MOPA) # Units # Units (no MOPA)
ADC XT1221 34 21 5 3
DAC XT1541 17 14 3 2
BIO XT1111 19 10 2 1

 



3113 - ADC

C1:PSL-126MOPA_126PWR
C1:PSL-126MOPA_DTMP
C1:PSL-126MOPA_LTMP
C1:PSL-126MOPA_DMON
C1:PSL-126MOPA_LMON
C1:PSL-126MOPA_CURMON
C1:PSL-126MOPA_DTEC
C1:PSL-126MOPA_LTEC
C1:PSL-126MOPA_CURMON2
C1:PSL-126MOPA_HTEMP
C1:PSL-126MOPA_HTEMPSET
C1:PSL-FSS_RFPDDC
C1:PSL-FSS_LODET
C1:PSL-FSS_FAST
C1:PSL-FSS_PCDRIVE
C1:PSL-FSS_MODET
C1:PSL-FSS_VCODETPWR
C1:PSL-FSS_TIDALOUT
C1:PSL-PMC_RFPDDC
C1:PSL-PMC_LODET
C1:PSL-PMC_PZT
C1:PSL-PMC_MODET


3123 - ADC (failed)

C1:PSL-126MOPA_AMPMON
C1:PSL-126MOPA_126MON
C1:PSL-FSS_RCTRANSPD
C1:PSL-FSS_MINCOMEAS
C1:PSL-FSS_RMTEMP
C1:PSL-FSS_RCTEMP
C1:PSL-FSS_MIXERM
C1:PSL-FSS_SLOWM
C1:PSL-FSS_TIDALINPUT
C1:PSL-PMC_PMCTRANSPD
C1:PSL-PMC_PMCERR
C1:PSL-PPKTP_TEMP


4116 - DAC

C1:PSL-126MOPA_126CURADJ
C1:PSL-126MOPA_DCAMP
C1:PSL-126MOPA_DCAMP-
C1:PSL-FSS_INOFFSET
C1:PSL-FSS_MGAIN
C1:PSL-FSS_FASTGAIN
C1:PSL-FSS_PHCON
C1:PSL-FSS_RFADJ
C1:PSL-FSS_SLOWDC
C1:PSL-FSS_VCOMODLEVEL
C1:PSL-FSS_TIDAL
C1:PSL-FSS_TIDALSET
C1:PSL-PMC_GAIN
C1:PSL-PMC_INOFFSET
C1:PSL-PMC_PHCON
C1:PSL-PMC_RFADJ
C1:PSL-PMC_RAMP


XVME-210 - Binary Input

C1:PSL-126MOPA_FAULT
C1:PSL-126MOPA_INTERLOCK
C1:PSL-126MOPA_SHUTTER
C1:PSL-126MOPA_126LASE
C1:PSL-126MOPA_AMPON


XVME-220 - Binary Output

C1:PSL-126MOPA_126NE
C1:PSL-126MOPA_126STANDBY
C1:PSL-126MOPA_SHUTOPENEX
C1:PSL-126MOPA_STANDBY
C1:PSL-FSS_SW1
C1:PSL-FSS_SW2
C1:PSL-FSS_FASTSWEEP
C1:PSL-FSS_PHFLIP
C1:PSL-FSS_VCOTESTSW
C1:PSL-FSS_VCOWIDESW
C1:PSL-PMC_SW1
C1:PSL-PMC_SW2
C1:PSL-PMC_PHFLIP
C1:PSL-PMC_BLANK

  15253   Wed Mar 4 22:38:31 2020 JonUpdatePSLc1psl communications problem resolved

I investigated the problem reported earlier today with the BIO1 channels. By logging the systemd messages generated when the IOC starts, I was immediately able to determine that the problem was not limited to BIO1. The modbus communications were failing for several other units as well.

Because some in-situ rewiring of a handful of channels had recently been done (more on this soon), I initially suspected that one of the Acromags had been damaged in the process. However, removing BIO1 (or other non-communicating modules) did not restore communications with the rest of the modules. To test whether the chassis was the source of the problem at all, we set up a fresh ADC (new out of the package) and directly connected it to the secondary Ethernet interface of c1psl. With only the one new ADC connected, the modbus IOC failed in exactly the same way.

To confirm that the new ADC did in fact work, we connected it to c1auxex in the same configuration. The unit worked fine connected to c1auxex. This established that the source of the problem was the c1psl host. After some extensive debugging, I traced the problem to a pre-execution script (part of the modbus IOC systemd service) which resets the secondary network interface (the one connected to the Acromag chassis) prior to launching the IOC. This was to ensure the secondary interface always had the correct IP address. It appears this reset was somehow creating a race condition that allowed the modbus initializations (first communications with the Acromags) to sometimes start before the network interface had actually come back up.

I still don't understand how this was happening, or why the pre script worked just fine up until yesterday, but eliminating the network interface reset fixes the problem in 100% of the trials we ran. Unfortunately we lost the entire day to debugging this problem, so the final round of testing is still to be completed. We plan to pick it back up tomorrow afternoon.

  14817   Tue Jul 30 09:13:31 2019 gautamUpdatePSLc1psl keyed, Agilent setup cleared
  1. IMC would not lock. c1psl EPICS channels were unresponsive. I keyed the crate and went through the usual burtrestore/PMC-relocking dance.
  2. While at 1X2, I decided to take this opportunity to clean up the AG4395 setup that has been setup there unused for several weeks now.
    • Unplugged the active probe connected via BNC-T connector to the mixer IF output.
    • Noticed that the active probe (S/N 2850J01450) did not have it's power connection connected. According to the manual, this is bad. I don't know if the probe is damaged or not.
    • Moved the AG4395 cart out of the way so that there is a little more room around 1X1/1X2.
  15184   Mon Feb 3 15:22:39 2020 JonUpdatePSLc1psl progress/Acromag ADC grounding

I tested the c1psl AO channels on the electronics bench on Friday. While I found all the wiring to be correct, some of the channels exhibited excess noise with all appearances of a grounding problem.

Today Jordan, Gautam, and I investigated this further. It is indeed a grounding problem, but actually with the Acromag ADCs. The Acromag DAC outputs are single-ended (return is grounded), so (for the purpose of a loopback test) I would expect to leave the ADC inputs ungrounded. This is the configuration I tested Friday. Today we also tested driving the ADC with a floating source. The ADC noise behavior is exactly the same, whether the source end is grounded or not.

However, grounding the minus pin of the ADC channel eliminates the noise. We don't understand why this seems to be required irrespective of the driving source, so there something we're missing about the ADC design. As it turns out, this same fix was made to the AI channels of the previously-upgraded Acromag machines. I know Chub and I had to do this for the AI channels of c1vac, but at the time we thought the source grounding was causing the issue. However, today Jordan and I looked inside c1iscaux, which Chub wired, and confirmed that its AI channels are wired in the same way.

So in any case, Jordan is grounding the c1psl AI channels in the same way as c1iscaux. Once this is done, we'll continue with the bench testing tomorrow.

gautam: here are my notes about this issue when i was doing the c1iscaux testing. As I note there, "previously-upgraded Acromag machines" in the plural may be a bit of a stretch - I have no idea what the grounding situation is in c1susaux / c1auxex for example.

  15115   Fri Jan 10 14:21:19 2020 YehonathanUpdatePSLc1psl reboot

PSL controls on the sitemap went blank. Rebooted c1psl. PSL screens seem normal again.

  1976   Tue Sep 8 19:30:33 2009 ranaUpdatePSLc1psl rebooted for new RCPID database settings

The RC thermal PID is now controllable from its own MEDM screen which is reachable from the FSS screen. The slowpid.db and psl.db have been modified to add these records and all seems to be working fine.

Also, I've attached the c1psl startup output that we got on the terminal. This is just for posterity.

I'm also done tuning the PID for now. Using Kp = -1.0, Ki = -0.01, and Kd = 0, the can servo now has a time constant of ~10 minutes and good damping as can be seen in the StripTool snap below. These values are also now in the saverestore.req so hopefully its fully commissioned.

I bet that its much better now than the MINCO at holding against the 24 hour cycle and can nicely handle impulses (like when Steve scans the table). Lets revisit this in a week to see if it requires more tuning.

  1983   Thu Sep 10 18:25:15 2009 ranaUpdatePSLc1psl rebooted for new RCPID database settings

I added a new database record (C1:PSL-FSS_RCPID_SETPOINT) to allow for changing of the RC setpoint while the loop is on. This will enable us to step the can's temperature and see the result in the NRPO's SLOWDC.

 

  15231   Thu Feb 27 17:50:36 2020 gautamUpdatePSLc1psl setup setup

[many people]

in prep for the install tomorrow, we did the following:

  • Install the c1psl Supermicro in the 1X2 rack (Attachment 1). To make room we removed the anti-image filter and mounted it on the OMC rack.
  • Set up a local workstation (monitor+mouse+keyboard) for the Supermicro so we can do some local testing (Attachment 2).
  • Clear up the immediate area around the 1X1/1X2 rack, setup a cart for the Acromag.
  • Make sure there are sufficient adaptor boards cables (DB37, DB15, DB9, DB25, ethernet) etc available at the cart.
  • Label cables, connect on Acromag chassis end (Attachment 3).
  • Keep some large (A3) printouts of the channel mapping handy by the cart.
  • made sure we have open fuse-able DIN rail connectors for +/-15 V DC and +/-24 V DC for the Acromag box (we are waiting on some thinner gauge cabling for the 24V supply, once that arrives, we will power the box from the Sorensens. For now, they are powered by bench supplies on the cart).
  • made sure c1psl1 (still this name for the Supermicro) is ssh-able.

Barring objections, tomorrow (Friday 28 Feb 2020) morning I will commence the switch (I still want to work on the IFO tonight).

  15234   Fri Feb 28 08:05:22 2020 gautamUpdatePSLc1psl setup setup

And so it begins.

Quote:

Barring objections, tomorrow (Friday 28 Feb 2020) morning I will commence the switch

  15235   Fri Feb 28 10:04:41 2020 gautamUpdatePSLc1psl setup setup

Summary:

There are several problems evident already.

  1. Several EPICS database entries were missing. WTF.
  2. After fixing the missing entries, the PMC could be locked. However, the IMC could not be locked.
  3. I think the FSS Interface card is not configured correctly.

For now, I've returned the old c1psl connections, the PMC and IMC are both locked. Need to do some debugging on the bench.

  15239   Mon Mar 2 16:35:12 2020 gautamUpdateCDSc1psl test status

Channel list with test status
== Test Status ==

[done] Lock PMC and IMC
[done] IMC Servo board test
[done] IMC LO Det Mon channel check
[0th order] WFS quadrant DC mon
[none] WFS I/F monitors
[0th order] WFS attenuators
[none] IOO QPD channels
[done] FSS readbacks 
[done] PMC readbacks


Some more detailed elogs about the individual tests will follow.

Basically, I have characterized the IMC Servo board in detail. The summary finding is that the IN2 (=AO gain) slider needs to be investigated. 

All other channels need to be verified in a more thorough fashion than my basic checks which were just to guarantee the core interferometer functionality, which is important to me.

  12849   Thu Feb 23 15:48:43 2017 johannesUpdateComputersc1psl un-bootable

Using the PDA520 detector on the AS port I tried to get some better estimates for the round-trip loss in both arms. While setting up the measurement I noticed some strange output on the scope I'm using to measure the amount of reflected light.

The interferometer was aligned using the dither scripts for both arms. Then, ITMY was majorly misaligned in pitch AND yaw such that the PD reading did not change anymore. Thus, only light reflected from the XARM was incident of the AS PD. The scope was showing strange oscillations (Channel 2 is the AS PD signal):

For the measurement we compare the DC level of the reflection with the ETM aligned (and the arm locked) vs a misaligned ETM (only ITM reflection). This ringing could be observed in both states, and was qualitatively reproducible with the other arm. It did not show up in the MC or ARM transmission. I found that changing the pitch of the 'active' ITM (=of the arm under investigation) either way by just a couple of ticks made it go away and settle roughly at the lower bound of the oscillation:

In this configuration the PD output follows the mode cleaner transmission (Channel 3 in the screen caps) quite well, but we can't take the differential measurement like this, because it is impossible to align and lock the arm but them misalign the ITM. Moving the respective other ITM for potential secondary beams did not seem to have an obvious effect, although I do suspect a ghost/secondary beam to be the culprit for this. I moved the PDA520 on the optical table but didn't see a change in the ringing amplitude. I do need to check the PD reflection though.

Obviously it will be hard to determine the arm loss this way, but for now I used the averaging function of the scope to get rid of the ringing. What this gave me was:
(16 +/- 9) ppm losses in the x-arm and (-18+/-8) ppm losses in the y-arm

The negative loss obviously makes little sense, and even the x-arm number seems a little too low to be true. I strongly suspect the ringing is responsible and wanted to investigate this further today, but a problem with c1psl came up that shut down all work on this until it is fixed:

I found the PMC unlocked this morning and c1psl (amongst other slow machines) was unresponsive, so I power-cycled them. All except c1psl came back to normal operation. The PMC transmission, as recorded by c1psl,  shows that it has been down for several days:

Repeated attempts to reset and/or power-cycle it by Gautam and myself could not bring it back. The fail indicator LED of a single daughter card (the DOUT XVME-212) turns off after reboot, all others stay lit. The sysfail LED on the crate is also on, but according to elog 10015 this is 'normal'. I'm following up that post's elog tree to monitor the startup of c1psl through its system console via a serial connection to find out what is wrong.

  12850   Thu Feb 23 18:52:53 2017 ranaUpdateComputersc1psl un-bootable

The fringes seen on the oscope are mostly likely due to the interference from multiple light beams. If there are laser beams hitting mirrors which are moving, the resultant interference signal could be modulated at several Hertz, if, for example, one of the mirrors had its local damping disabled.

  12851   Thu Feb 23 19:44:48 2017 johannesUpdateComputersc1psl un-bootable

Yes, that was one of the things that I wanted to look into. One thing Gautam and I did that I didn't mention was to reconnect the SRM satellite box and move the optic around a bit, which didn't change anything. Once the c1psl problem is fixed we'll resume with that.

Quote:

The fringes seen on the oscope are mostly likely due to the interference from multiple light beams. If there are laser beams hitting mirrors which are moving, the resultant interference signal could be modulated at several Hertz, if, for example, one of the mirrors had its local damping disabled.

 

Speaking of which:

Using one of the grey RJ45 to D-Sub cables with an RS232 to USB adapter I was able to capture the startup log of c1psl (using the usb camera windows laptop). I also logged the startup of the "healthy" c1aux, both are attached. c1psl stalls at a point were c1aux starts testing for present vme modules and doesn't continue, however is not strictly hung up, as it still registers to the logger when external login attempts via telnet occur. The telnet client simply reports that the "shell is locked" and exits. It is possible that one of the daughter cards causes this. This seems to happen after iocInit is called by the startup script at /cvs/cds/caltech/target/c1psl/startup.cmd, as it never gets to the next item "coreRelease()". Gautam and I were trying to find out what happends inside iocInit, but it's not clear to us at this point from where it is even called. iocInit.c and compiled binaries exist in several places on the shared drive. However, all belong to R3.14.x epics releases, while the logfile states that the R3.12.2 epics core is used when iocInit is called.

Next we'll interrupt the autoboot procedure and try to work with the machine directly.

  12854   Tue Feb 28 01:28:52 2017 johannesUpdateComputersc1psl un-bootable

It turned out the 'ringing' was caused by the respective other ETM still being aligned. For these reflection measurements both test masses of the other arm need to be misaligned. For the ETM it's sufficient to use the Misalign button in the medm screens, while the ITM has to be manually misaligned to move the reflected beam off the PD.

I did another round of armloss measurements today. I encountered some problems along the way

  • Some time today (around 6pm) most of the front end models had crashed and needed to be restarted GV: actually it was only the models on c1lsc that had crashed. I noticed this on Friday too.
  • ETMX keeps getting kicked up seemingly randomly. However, it settles fast into it's original position.

General Stuff:

  • Oscilloscope should sample both MC power (from MC2 transmitted beam) and AS signal
  • Channel data can only be loaded from the scope one channel at a time, so 'stop' scope acquisition and then grab the relevant channels individually
  • Averaging needs to be restarted everytime the mirrors are moved triggering stop and run remotely via the http interface scripts does this.

Procedure:

  1.     Run LSC Offsets
  2.     With the PSL shutter closed measure scope channel dark offsets, then open shutter
  3.     Align all four test masses with dithering to make sure the IFO alignment is in a known state
  4.     Pick an arm to measure
  5.     Turn the other arm's dither alignment off
  6.     'Misalign' that arm's ETM using medm screen button
  7.     Misalign that arm's ITM manually after disabling its OpLev servos looking at the AS port camera and make sure it doesn't hit the PD anymore.
  8.     Disable dithering for primary arm
  9.     Record MC and AS time series from (paused) scope
  10.     Misalign primary ETM
  11.     Repeat scope data recording

Each pair of readings gives the reflected power at the AS port normalized to the IMC stored power:

\widehat{P}=\frac{P_{AS}-\overline{P}_{AS}^\mathrm{dark}}{P_{MC}-\overline{P}_{MC}^\mathrm{dark}}

which is then averaged. The loss is calculated from the ratio of reflected power in the locked (L) vs misaligned (M) state from

\mathcal{L}=\frac{T_1}{4\gamma}\left[1-\frac{\overline{\widehat{P}_L}}{\overline{\widehat{P}_M}} +T_1\right ]-T_2

Acquiring data this way yielded P_L/P_M=1.00507 +/- 0.00087 for the X arm and P_L/P_M=1.00753 +/- 0.00095 for the Y arm. With \gamma_x=0.832 and \gamma_x=0.875 (from m1=0.179, m2=0.226 and 91.2% and 86.7% mode matching in X and Y arm, respectively) this yields round trip losses of:

\mathcal{L}_X=21\pm4\,\mathrm{ppm}  and  \mathcal{L}_Y=13\pm4\,\mathrm{ppm}, which is assuming a generalized 1% error in test mass transmissivities and modulation indices. As we discussed, this seems a little too good to be true, but at least the numbers are not negative.

  14455   Thu Feb 14 23:14:12 2019 gautamUpdateCDSc1rfm errors

The pressure is still 2e-4 torr according to CC1 so I thought I'd give ASS debugging a go tonight. But the arm transmission signal isn't coming through to the LSC model from the end PDs - so a resurfacing of this problem. Rebooting the sender model, c1scy, did not fix the problem. Moreover, c1susaux is dead. The last time I rebooted it, ITMY got stuck so I'm not going to attempt a revival tonight.

  15240   Mon Mar 2 19:32:41 2020 gautamUpdateCDSc1rfm errors

Had to reboot both end machines and the c1rfm model to get the TRX and TRY signals to the LSC models. Now both arms can be locked using POX/POY respectively.

  14457   Fri Feb 15 15:22:08 2019 gautamUpdateCDSc1rfm errors persist

I restarted c1scyc1rfm (so both sender and receiver models were cycled) and power-cycled the c1iscey and c1sus machines. The TRY PD is certainly seeing light - it is just not getting piped over to c1rfm. dmesg doesn't give any clues. I'm out of ideas.

P.S. The new reality seems to be that getting ITMY stuck in the event of a c1susaux reboot is inevitable. As is the practise for ITMX, I tried slowly ramping the PIT and YAW biases to 0 slowly - but in the process of ramping YAW to 0, the optic got stuck. I am ramping in steps of 0.1 (in units of the PIT/YAW sliders, waiting ~3 seconds between steps), I guess I can try ramping even more slowly.

Update: I power cycled the physical RFM switch. This necessitated reboot of all vertex FEs. But seems like things are back to normal now...

Note: to unstick ITMY, seems like the best approach is:

  1. Jiggle bias until SIDE shadow sensor is on average above it's half-light level. This is the critical step. A bias of +20000 cts on the fast SIDE output seems to help.
  2. Set YAW bias to -10, ramp down the BIAS in steps of 0.1, watching shadow sensor levels to ensure optic doesn't get stuck again.
  3. Hope for the best. Iterate if necessary.
Quote:

The pressure is still 2e-4 torr according to CC1 so I thought I'd give ASS debugging a go tonight. But the arm transmission signal isn't coming through to the LSC model from the end PDs - so a resurfacing of this problem. Rebooting the sender model, c1scy, did not fix the problem. Moreover, c1susaux is dead. The last time I rebooted it, ITMY got stuck so I'm not going to attempt a revival tonight.

  15920   Mon Mar 15 20:22:01 2021 gautamUpdateASCc1rfm model restarted

On Friday, I felt that the ASC performance when the PRFPMI was locked was not as good as it used to be, so I looked into the situation a bit more. As part of my ASC model revamp in December, I made a bunch of changes to the signal routing, and my suspicion was that the control signals weren't even reaching the ETMs. My log says that I recompiled and reinstalled the c1rfm model (used to pipe the ASC control signals to the ETMs), and indeed, the file was modified on Dec 21. But for whatever reason, the C1RFM.ini (=Dolphin receiver since the ASC control signals are sent to this model over the Dolphin network from the c1ioo machine which hosts the C1:ASC- namespace, and RFM sender to the ETMs, but this path already existed) file never picked up the new channels. Today, I recompiled, re-installed, and restarted the models, and confirmed that the control signals actually make it to the ETMs. So now we can have the QPD-based ASC loops engaged once again for the PRFPMI lock. The CDS system did not crash 🎉 . See Attachments #1-3.

I checked the loop performance in the POX/POY locked config by first deliberately misaligning the ETMs, and then engagin the loops - seems to work (Attachment #4). The loop shapes have to be tweaked a bit and I didn't engage the integrators, hence the DC pointing wasn't recovered. Also, added a line to the script that turns the ASC loops on to set limits for all the loops - in the testing process, one of the loops ran away and I tripped the ETMY watchdog. It has since been recovered. I SDFed a limit of 100cts just to be on the conservative side for model reboot situations - the value in the script can be raised/lowered as deemed necessary (sorry, I don't know the cts-->urad number off the top of my head).

But the hope is this improves the power buildup, and provides stability so that I can begin to commission the AS WFS system a bit.

  11883   Tue Dec 15 11:22:53 2015 gautamUpdateCDSc1scx and c1asx crashed

I noticed what I thought was excessive movement of the beam spot on ITMX and ETMX on the control room monitors, and when I checked the CDS FE status overview MEDM screen, I saw that c1scx and c1asx had crashed. I ssh-ed into c1iscex and restarted both models, and then restarted fb as well. However, the DAQ-DCO_C1SCX_STATUS indicator remains red even after restarting fb (see attached screenshot). I am not sure how to fix this so I am leaving it as is for now, and the X arm looks to have settled down.

  7008   Mon Jul 23 18:57:52 2012 JamieUpdateCDSc1scx and c1scy models recompiled and restarted

After the changes listed in 7005 and 7007, I have rebuilt, installed, and restarted the c1scx and c1scy models.  Everything seems to have come back up ok.

Running into some daqd troubles because of a change to c1ioo, but will report on the new ALS channels when I can.

  6436   Thu Mar 22 16:45:06 2012 kiwamuUpdateCDSc1scx and c1scy not properly running

It seems that neither c1scx nor c1scy is working properly as their ADC counts are showing digital-zeros.

However the IOPs, c1gcx and c1gcy look running fine, and also the IOPs seem successfully recognizing the ADCs according to dmesg.

Also there is one more confusing fact : c1scx and c1scy are synchronizing to the timing signal somehow.

I restarted the c1scx front end model to see if this helps, but unfortunately it didn't work.

As this is not the top priority concern for now, I am leaving them as they are now with the watchgods off.

(I may try hardware rebooting them in this evening)

Quote from #6434

The power was turned back on at 4pm It took some time for Suresh to restart the computers. We have damping but things are not perfect yet. Auto BURTH did not work well.

 

  6438   Thu Mar 22 17:41:15 2012 sureshUpdateCDSc1scx and c1scy not properly running

Quote:

It seems that neither c1scx nor c1scy is working properly as their ADC counts are showing digital-zeros.

Quote from #6434

The power was turned back on at 4pm It took some time for Suresh to restart the computers. We have damping but things are not perfect yet. Auto BURTH did not work well.

 When Steve and I restarted the c1iscex and c1iscey computers after the power shutdown, the models within them did not start-up automatically.  I had to start them manually from a terminal in the control room. 

I also tried rebooting the FB a couple of times.  Did not make any difference.

Manually starting the c1x05, c1scy and c1x01, c1scx models (with the Burt Restore button ON) did not resolve the issue of zeros in the epics screens.  though it did re-establish timing. 

  6439   Thu Mar 22 23:43:56 2012 KojiUpdateCDSc1scx and c1scy not properly running

Did you guys checked if the simplant switch is set to "REAL WORLD" mode?

Edit by KI:

Bingo ! The input signals were bypassed to the simplant. I switched the simplant settings to REAL WORLD and now both end suspensions are working fine.

  5535   Sat Sep 24 01:38:14 2011 kiwamuUpdateCDSc1scx and c1x01 restarted

[Koji / Kiwamu]

 The c1scx and c1x01 realtime processes became frozen. We restarted them around 1:30 by sshing and running the kill/start scripts.

  6175   Fri Jan 6 01:00:56 2012 kiwamuUpdateCDSc1scx out of sync

Both the c1scx and its IOP realtime processes became out of sync.

Initially I found that the c1scx didn't show any ADC signals, though the sync sign was green.

Then I software-rebooted the c1iscex machine and then it became out of sync.

For tonight this is fine because I am concentrating on the central part anyway.