ID |
Date |
Author |
Type |
Category |
Subject |
10604
|
Mon Oct 13 21:59:47 2014 |
rana | Update | Computer Scripts / Programs | Which side of optical spring are we on? (No progress) |
Since no one was here, I started the Ubuntu 10 - 12 upgrade on Rossa. It didn't run at first because it wanted to remove 'update-manager-kde' even though it was on the blacklist. I removed it from the command line and now its running. Allegra, OTOH, refuses to upgrade. Someone please ask Diego to wipe it and then install Ubuntu 12 LTS on there in the morning...its a good way to learn the Martian CDS setup. |
10612
|
Wed Oct 15 19:56:38 2014 |
Jenne | Update | LSC | Which side of optical spring are we on? Meas vs Model |
I have plotted measured data from last night (elog 10607) with a version of the result from Rana's simulink CARM loop model (elog 10593).
The measured data that was taken last night (open circles in plots) is with an injection into MC2 position, and I'm reading out TRX. This is for the negative side of the digital CARM offset, which is the one that we can only get to arm powers of 5ish.
The modeled data (solid lines in plots) is derived from what Rana has been plotting the last few days, but it's not quite identical. I added another excitation point to the simulink model at the same place as the "CARM OUT" measurement point. This is to match the fact that the measured transfer functions were taken by driving MC2. I then asked matlab to give me the transfer function between this new excitation point (CARM CTRL point) and the IN1 point of the loop, which should be equivalent to our TRX_OUT. So, I believe that what I'm plotting is equivalent to TRX/MC2. The difference between the 2 plots is just that one uses the modeled spring-side optical response, and the other uses the modeled antispring-side response.


I have zoomed the X-axis of these plots to be between 30 Hz - 3 kHz, which is the range that we had coherence of better than 0.8ish last night in the measurements. The modeled data is all given the same scale factor (even between plots), and is set so that the lowest arm power traces (pink) line up around 150 Hz.
I conclude from these plots that we still don't know what side of the CARM resonance we are on.
I have not plotted the measurements from the positive side of the digital CARM offset, because those transfer functions were to sqrtInvTRX, not plain TRX, whereas the model only is for plain TRX. There should only be an overall gain difference between them though, no phase difference. If you look at last night's data, you'll see that the positive side of the CARM offset measured phase has similar characteristics to the negative offset, i.e. the phase is not flat, but it is roughly flat in both modeled cases, so even with that data, I still say that we don't know what side of the CARM resonance we are on.
|
14759
|
Mon Jul 15 03:30:47 2019 |
Kruthi | Update | Calibration-Repair | White paper as a Lambertian scatterer |
I made some rough measurements, using the setup I had used for CCD calibration, to get an idea of how good of a Lambertian scatterer the white paper is. Following are the values I got:
Angle (degrees) |
Photodiode reading (V) |
Ps (W) |
BRDF (per str) |
% error |
12 |
0.864 |
2.54E-06 |
0.334 |
20.5 |
24 |
0.926 |
2.72E-06 |
0.439 |
19.0 |
30 |
1.581 |
4.65E-06 |
0.528 |
19.0 |
41 |
0.94 |
2.76E-06 |
0.473 |
19.8 |
49 |
0.545 |
1.60E-06 |
0.423 |
22.5 |
63 |
0.371 |
1.09E-06 |
0.475 |
28 |
Note: All the measurements are just rough ones and are prone to larger errors than estimated.
I also measured the transmittance of the white paper sample being used (it consists of 2 white papers wrapped together). It was around 0.002 |
Attachment 1: BRDF_paper.png
|
|
17584
|
Mon May 8 17:05:30 2023 |
Yehonathan | Update | BHD | Whitening TF measurements |
{Mayank, Yehonathan}
We measured today the TFs of the whitening boards. We measured in particular REFL11 I/Q and AS55 I/Q channels using SR785.
There seems to be an issue with turning on whitening gain bigger than 18dB. In all our measurements, when the whitening filter was off the TF was flat and had the right gain. However, when we turned the whitening on, the measured TFs for gains higher than 18 dB would like exactly like as if the whitening gain was 18 dB. This happened in all channels that were measured and across two separate whitening filter boards.
Also, it was hard to measure both low and high-frequency parts of the TFs when the gain was high. The gain difference should be normally 40 dB but for higher gains it seems smaller. We verified that at higher gain level the high-frequency response was dependant on the ecxitation level meaning we had some saturation there.
|
Attachment 1: Whitening_TFs_REFL11_I.pdf
|
|
Attachment 2: Whitening_TFs_REFL11_Q.pdf
|
|
Attachment 3: Whitening_TFs_AS55_I.pdf
|
|
Attachment 4: Whitening_TFs_AS55_Q.pdf
|
|
17585
|
Tue May 9 11:32:04 2023 |
Yehonathan | Update | BHD | Whitening TF measurements |
We forgot to take a reference TF measurement by looping the SR785 on itself using the same BNC cables used for the actual measurement. I took this measurement today (attachment 1). As can be seen, there is a significant delay in the SR785 + cables themselves.
I also retook some measurements on the AS5_I whitening channel for various gains. Being careful with the excitation level and the channel range on the SR785 to avoid saturation I was also able to see low-frequency gains higher than 18dB so that problem is gone too. The results are shown in attachment 2 with the reference phase subtracted from the measurements. |
Attachment 1: Reference_TF.pdf
|
|
Attachment 2: Whitening_TFs_AS55_I.pdf
|
|
15531
|
Mon Aug 17 23:36:10 2020 |
gautam | Update | ALS | Whitening and ALS noise |
I finally managed to install a differential-receiving whitening board in 1Y2 - 4 channels are available at the moment. As I claimed, one stage of 15:150 Hz z:p whitening does improve the ALS noise a little, see Attachment #1. While the RMS (from 1kHz-0.5 Hz) does go down by ~10 Hz, this isn't really going to make any dramatic improvement to the 40m lock acquisiton. Now we're really sitting on the unsuppressed EX laser noise above ~30 Hz. This measurement was taken with the arm cavities locked with POX/POY, and end lasers locked to the arm cavities with uPDH boxes as usual. This was just a test to confirm my suspicion, the whitening board is to be used for the air BHD channels, but when we get a few more stuffed, we can install it for the ALS channels too. |
Attachment 1: ALSimprovement.pdf
|
|
15533
|
Tue Aug 18 13:55:23 2020 |
rana | Update | ALS | Whitening and ALS noise |
No, there should be no unscheduled visits from any inspector, marshal, tech, or vendor. They all have to be escorted or they don't get in. If they have a problem with that, please give them my cell #.
For the ALS, in addition to the beat note spectrum, I think we need to know the loop gain use to feedback to the ETM to determine the true cavity length fluctuation. w/o ALS, the noise would be only due to the seismic noise, OSEM damping noise, and the IR-PDH residual. Those are all suppressed by the ALS loop, but then the ALS loop puts its sensing noise onto the cavity. So, if I'm thinking about this right, the ALS beat noise > 200 Hz doesn't matter so much to the CARM RMS. So the whitening seems to be doing good in the right spot, but we would like to have another boost in the green PDH to up the gain below ~300 Hz? |
15532
|
Mon Aug 17 23:41:50 2020 |
gautam | Update | BHD | Whitening and air BHD dark noise |
Summary:
With the chosen transimpedance of 300 ohms, in order to be able to see the shot noise of 10 mW of light in the digitized data streams, we'd need all 3 stages of whitening. If we want to be shot noise limited with 1 mW of LO light, we'd need to increase said transimpedance I think.
Details:
The measurements were taken with
- No light incident on the DCPDs.
- The flat whitening gain was set to 0 dB.
- Whitening engaged sequentially, stage by stage, shown as (Blue, Red, Orange and Green) curves corresponding to (0, 1, 2, 3) stages of whitening.
Of course, it's unlikely we're going to be shot noise limited for any configuration in the short run. But this was also a test of
- My soldering.
- Change of whitening corner frequencies.
- Test of the overall whitening board assembly.
All 3 tests passed. |
Attachment 1: BHD_whitening.pdf
|
|
16972
|
Tue Jul 5 20:05:06 2022 |
Tomislav | Update | Electronics | Whitening electronics noise |
For whitening electronics noise for WFS1, I get (attachment). This doesn't seem right, right? |
Attachment 1: whitening_noises.png
|
|
13563
|
Sat Jan 20 01:20:37 2018 |
gautam | Update | Electronics | Whitening filter D990694 |
We use D990694 in various places. Today, Rana alerted me to an important consideration to be kept in mind when we use this board, which I found quite interesting. I still don't understand the problem at the BJT level, but I think one can appreciate the problem without going to the transistor design of the LT1125. I'm attaching an annotated schematic of the whitening section in question. If the following assumptions are valid, then I think my picture is valid.
- The switch used to bypass the various whitening gain stages, namely the ADG333ABR, has infinite impedance in the "OFF" state, such that when the 24dB gain stage is bypassed, U28A (or in general one of the 4 quad op-amps) is forced to drive it's output voltage across 1.0665 kohms of resistance.
- The individual LT1125 Op Amps can drive a maximum of 30mA of current.
Then, as one can see in the attached schematic, when we set the gain of any input to <24dB, we must ensure that the input voltage is less than approximately 2V. Otherwise, by asking too much of the first stage op-amp in the quad IC LT1125, we may be messign around with all the 4 op amps in the quad! Even the 0dB setting is not immune to this problem, as it uses one of the 4 op amps.
I don't think the usual rules of calculating the gain of a non-inverting amplifier (G = 1 + Rf/Ri) remain valid even when the op-amp is forced to drive more output current than it can, and I don't have a way to quantify the possible interference between the 4 op amps in the quad - but does this seem like a valid conclusion? If so, we must check signal levels of various LSC signals. AS55 signals currently have the 0dB gain setting - I had turned this down from 6dB some months ago, because it seemed like the ADC was saturating at the 6dB gain setting, which suggests that the input voltage is ceratinly > 2V, and AS55_Q is what is used for MICH control in the DRMI. All of my noise budgeting work over the last few months used this setting, I wonder if they are all invalid 
Now that I think about this a bit more - this problem shouldn't be significant for the usual LSC degrees of freedom when in lock, as the huge DC gain of the loop should squish large DC values of the error signals, and so there shouldn't be any danger of overloading the LT1125. But I don't know if we are being hurt by this effect when flashing through resonances, when the PDH horn-to-horn voltage can be quite high (which is in principle a good thing?). I don't know if there is any "hysterisis" effect where the overloaded quad IC has some relaxation time before it returns to normal operation, and if we are being limited in our ability to catch lock because if this effect.
The concerns remain valid for th ALS demodulated error signals though, for which the signals will remain large throughout. |
Attachment 1: whiteningBoardLimitations.pdf
|
|
13564
|
Sat Jan 20 15:57:11 2018 |
rana | Update | Electronics | Whitening filter D990694 |
this is the note from Hartmut Grote on this topic from 2004 |
13568
|
Tue Jan 23 01:33:23 2018 |
gautam | Update | Electronics | Whitening filter D990694 |
After discussing with Koji, we looked at the aLIGO incarnation of this board. Interestingly, it too has a similar topology of 4 switchable gain stages with gains of 24, 12, 6 and 3dB. The main differences are that they use single Op27 ICs instead of the quad LT1125s, and also, they use a different combination of feedback resistors to realize the various gains.
We considered upping the feedback resistance (R15, R143) on the 24dB gain stage of our boards from (1k, 66.5ohms) to (3k, 200ohms) as on the aLIGO boards - but this doesn't really help? Because KCL demands that the same current flow in R15 and R143, and so the output Vsat of the op amp and its max current driving capabilities in combination determine if the inverting input can follow the non inverting input?
As Hartmut points out in his note, he was able to access the full range of ADC voltages when the gain was set to 3dB, despite the fact that the LT1125 was still getting internally saturated. Operating with minimum 24dB whitening gain doesn't really solve the problem either because the problem just gets shifted to the next gain stage in the chain, and we still have saturation. I also don't have a feeling for how much differential voltage these LT1125s can sustain before they are damaged - I guess the planned THD check will reveal if they are okay or not.
It seems to me like the only way to truly fix this problem of one stage saturating and screwing up the others is to use single Op27s (or equivalent) in place of the quad LT1125s. The aLIGO design also has a series resistance to the non-inverting input - this can help prevent current overdraw from the previous stage (due to a lowered input impedance of the OpAmp - but I wonder how low this can go?).
|
13572
|
Wed Jan 24 00:48:47 2018 |
gautam | Update | Electronics | Whitening filter D990694 |
I plan to do some characterization of this problem. The plan is to use THD as a metric for whether we are having hidden saturations. Pg 9 of the LT1125 datasheet tells us what fraction of THD to expect. I will use one of the several unused DAC channels available at the LSC rack to drive a 100Hz sine wave into one of the inputs of the whitening chassis, and measure the THD up to a reasonable harmonic number (will probably be set by the ADC noise) for (i) various whitening gain settings and (ii) various input signal amplitudes.
The motivation is to attempt to quantify the problem better:
- How bad is it to have one or more of the OpAmps in the quad IC either saturated to its voltage rails, or max output current?
- Can we reproduce Hartmut's observations?
- Are the OpAmps already irreversibly damaged because of extended abuse?
Then we can decide what, if anything, to do about this issue. |
17582
|
Wed May 3 18:40:50 2023 |
Yehonathan | Update | BHD | Whitening noises measurements |
{Mayank, Yehonathan}
We measured the noise at the WF1 (REFL11) and WF2 (AS55) boards at the LSC rack with and without whitening filter. We switch the filter on and off by switching off and on the unwhitening in the PDs filter bank.
Attachment 1 shows the measurements.
Attachment 2 shows the ratio between the noise with and without whitening filter. I also plot the inverse of the unwhitening MEDM filter (all the unWhite filters were the same). I tune the gain of that filter to match the ratio of the AS55 whitening noises.
This is because I couldn't match the ratio of the REFL11 noises.
Moreover, the overall gain doesn't make sense to me. AS55 whitening has a gain of 24db and REFL11 has a gain of 18db. I'm not entirely sure where these values should show up. Also seems like REFL11 whitening has more gain than AS55 whitening. Will have to investigate more tomorrow. |
Attachment 1: Whitening_noises.pdf
|
|
Attachment 2: Whitening_noises_on_off_ratios.pdf
|
|
3790
|
Tue Oct 26 22:57:37 2010 |
Jenne | Configuration | Computers | Why doesn't DTT work?!? |
DTT has only SUS and "X02" channels under C1 in the drop down channel selection menu. Basically, we can't measure any fast channels with DTT. I keep getting the error: "Unable to select testpoints." Sadface.
Similar things are true for DataViewer. The same limited number of fast channels, and no data found:
Server error 13: no data found
datasrv: DataWriteRealtime failed: daq_send: Illegal seek
Is this a framebuilder problem? Is this something that the CDS team has on the to-do list? |
3793
|
Wed Oct 27 10:53:03 2010 |
josephb | Configuration | Computers | Why doesn't DTT work?!? |
Test points for the SUS channels should be there. They have been working previously this week. Possibly break down points include awgtpman, mx_streams, and the fb itself. I'll look into that.
As far as other fast channels, there are no other fast front ends running than the suspensions ones we have. Until additional channels get connected to the front ends and the models updated, those are the channels we have available. However I am working on getting c1ioo up and running, and we can try connecting in some PEM channels today to the c1sus front end's 4th ADC.
Edit:
I tried starting a fresh instance of the frame builder, but when I brought the old copy down, it left a pair of zombie or dead mx_stream processes running on c1sus . Basically c1mcs and c1rms were still running, while c1x02 and c1sus came down. I tried to kill the processes but this caused the c1sus machine to crash. In the past I've killed left over mx_stream processes running after the frame builder has gone down, but I've never seen them crash the computer. I'm unsure why this happened since we haven't done any updates of the code, just updated models and daq configuration files.
Quote: |
DTT has only SUS and "X02" channels under C1 in the drop down channel selection menu. Basically, we can't measure any fast channels with DTT. I keep getting the error: "Unable to select testpoints." Sadface.
Similar things are true for DataViewer. The same limited number of fast channels, and no data found:
Server error 13: no data found
datasrv: DataWriteRealtime failed: daq_send: Illegal seek
Is this a framebuilder problem? Is this something that the CDS team has on the to-do list?
|
|
4027
|
Wed Dec 8 14:46:19 2010 |
josephb, kiwamu | Update | CDS | Why the ETMX daq channels were not recorded last night |
When adding the ETMX DAQ channels using the daqconfig gui (located in /opt/rtcds/caltech/c1/scripts/) on C1SCX.ini, we forgot to set the acquire flag to 1 from 0.
So the frame builder was receiving the data, but not recording it.
We have since then added ETMX and the C1SCX.ini file to Yuta's useful "activateDAQ.py" script in /opt/rtcds/caltech/c1/chans/daq/, so that it now sets the sensor and SUSPOS like channels to be acquired at 2k when run. You still need to restart the frame builder (telnet fb 8087 and then shutdown) for these changes to take effect.
The script now also properly handles files which already have had channels activated, but not acquired. |
14829
|
Mon Aug 5 17:23:26 2019 |
gautam | Summary | Computers | WiFi Settings on asia |
The VEA laptop asia was configured to be able to connect to too many WiFi networks - it was getting conflicted in its default position at the vertex and trying to hop between networks, for some reason trying to connect to networks that had poor signal strength. I deleted all options from the known networks except 40MARS. Now the network connection seems much more stable and reliable. |
13905
|
Thu May 31 19:51:06 2018 |
Koji | Update | General | WiFi router firmware update / rebooting |
The model of our martian wifi router (NETGEAR R6400) was found in the FBI router list to be rebooted asociated with the malware "VPNFilter" issue.
I checked the attached devices and found bunch of (legit) devices blocked to access the wifi router. This is not an immediate problem as most of the packets do not go through the wifi router. But potentially a problem in some cases like Wifi enables GPIB adapters. So I marked them to be "allowed".
In this opprtunity, I have updated the firmware of the wifi router and this naturally involved rebooting of the device.
|
2664
|
Tue Mar 9 09:32:31 2010 |
Koji | Summary | General | Wideband measurement of Fast PZT response |
I have measured a wideband response of the fast PZT in the LWE NPRO 700mW in the Alberto's setup.
This is a basic measurement to determine how much phase modulation we can obtain by actuating the fast PZT,
primarily for the green locking experiment.
RESULT
- Above 200kHz, there are many resonances that screws up the phase.
- Modulation of 0.1rad can be easily obtained even at 10MHz if the modulation frequency is scanned.
- Change of the laser frequency in DC was observed depending on the modulation frequency.
i.e. At the resonance the laser frequency escaped from the RF spectrum analyzer.
This may induced by the heat dissipation in the PZT causing the temperature change of the crystal.
- Some concerns: Is there any undesired AM by the PZT modulation?
---
METHOD
1. Locked the PLL of for the PSL-NPRO beating at 20MHz.
2. Added the modulation signal to the NPRO PZT input.
I used the output of the network analyzer sweeping from 100kHz to 1MHz.
3. Measured the transfer function from the modulation input to the PLL error signal.
The PLL error is sensitive to the phase fluctuation of the laser. Found that the first resonance is at 200kHz.
The TF is not valid below 3kHz where the PLL suppresses the modulation.
4. Single frequency modulation: Disconnected the PLL setup.
Plug Marconi into the fast PZT input and modulate it at various frequencies.
Observing with the RF spectrum analyzer, I could see strong modulation below 1MHz.
It turned out later that the TF measurement missed the narrow peaks of the resonances due to the poor freq resolution.
Also the modulation depth varies frequency by frequency because of the resonances.
Scanned the frequency to have local maximum of the modulation depth. Adjusted the
modulation amplitude such that the carrier is suppressed (J0(m)=0 i.e. m~2.4). As I could not obtain
the carrier suppression at above 1MHz, the height of the carrier and the sidebands were measured.
The modulation frequency was swept from 100kHz to 10MHz.
5. Calibration. The TF measured has been calibrated using the modulation depth obtained at 100Hz,
where the resonance does not affect the response yet.
The responce of the PZT was ~10MHz/V below 30kHz. Looks not so strange although this valure is
little bit high from the spec (2MHz/V), and still higher than my previous experience at TAMA (5MHz/V).
Note that this calibration does not effect to the modulation depth of the single freq measurement as they are independent. |
Attachment 1: PZT_response.png
|
|
1105
|
Sun Nov 2 20:44:58 2008 |
rana | Update | ASS | Wiener Filter performance over 5 hours |
I took one 2 hour stretch of data to calculate a MISO Wiener filter to subtract the Ranger seismometer
and the 6 Wilcoxon accelerometers from the IOO-MC_L channel. I then used that static filter to calculate
the residual of the subtraction in 10 minute increments for 5 hours. The filter was calculated based upon
the first 2 hours of the stretch.
The MC lock stretch is from Oct 31 03:00 UTC (I think that we are -8 hours from UTC, but the DST confounds me).
So its from this past Thursday night.
I wrote a script (/users/rana/mat/wiener/mcl_comp.m) which takes the static filter and does a bunch of loops
of subtraction to get a residual power spectrum for each 10 minute interval.
In the attached PNG, you can see the result. The legend is in units of minutes from the initial t0 = 03:00 UTC.
BLACK-DASHED -- MCL spectrum before subtraction
I have also used dashed lines for some of the other traces where there is an excess above the unsubtracted data.
Other than those few times, the rest are all basically the same; this indicates that we can do fine with a very
slow adaptation time for the feed-forward filters-- a few hours of a time constant is not so bad.
After making the plot I noticed that the Ranger signal was totally railed and junky during this time.
This probably explains the terrible performance below 1 Hz (where are those Guralps?)
The second attached image is the same but in spectrogram form. |
Attachment 1: f.png
|
|
Attachment 2: f1.png
|
|
1111
|
Mon Nov 3 22:35:40 2008 |
rana | Update | ASS | Wiener Filter performance over 5 hours |
To speed up the Wiener filter work I defined a 256 Hz version of the original 16kHz IOO-MC_L signal. The
attached plots show that the FE decimation code works correctly in handling the anti-aliasing and
downsampling as expected. |
Attachment 1: DAQ.pdf
|
|
1112
|
Tue Nov 4 00:47:53 2008 |
rana | Update | ASS | Wiener Filter performance over 5 hours |
Same as before, but now with a working Ranger seismometer.
In the spectrogram, the color axis is now in dB. This is a whitened spectrogram, so 0 dB corresponds to
the average (median) subtraction. The color scale is adjusted so that the large transients are saturated
since they're not interesting; from the DV trend its some kind of huge glitch in the middle of the
night that saturated the MC1 accelerometers only (maybe a pump?).
The attached trend shows the 5 hours used in the analysis. |
Attachment 1: f2.png
|
|
Attachment 2: f.png
|
|
5102
|
Wed Aug 3 02:28:08 2011 |
Manuel, Ishwita | Update | WienerFiltering | Wiener Filtering in X-arm |
Wiener Filtering was applied on the data collected from the X-arm during the time: GPS time-996380715 (Aug 02, 2011. 21:25:00. PDT) to GPS time-996382215 (Aug 02, 2011. 21:50:00. PDT) for a duration of 1500 seconds. During this time the X-arm was locked, we checked it by acquiring data from channel C1:LSC-TRX_OUT_DQ .
The seismometers were near the beam splitter (guralp2) and near MC2 (guralp1).
Target data was obtained from channel C1:LSC-XARM_IN1_DQ.


Following graphs were obtained after applying the Wiener filter:
1.Seismic data acquired from Guralp1 (X and Y) and Guralp2 (X and Y) 2.Seismic data acquired from Guralp2 X 3.Seismic data acquired from Guralp2 Y
  
These graphs were obtained with srate = 2048 (sample rate) and N = 20000 (order of the filter).
Graph 1 is the best because the black (residual) line is below the red (target) line for low frequencies since we used seismic data from 4 channels. Graph 3 is the worst because we used seismic data from only one Y channel (Y axis of Guralp2) that is less related with the X-arm mirrors' motion since they are oriented orthogonally. |
854
|
Tue Aug 19 17:00:19 2008 |
Sharon | Update | | Wiener TF calibration - update |
This is an update for post 814
I added the calibration gains I got for the accelerometers (I realize I am just calibrating the accelerometers to themselves and this is not m/m exactly since we don't really know which accelerometer is doing exactly what we want it to do. However, since we are talking on relative small numbers, this shouldn't really change much).
I also added another missing gain for the seismometer. Rana has previously installed a 4300 ohm resistor in the seismometer, which changed the gain to 4300/(4300+5000) = 0.46 (this is from the manual). Moreover, there is a gain of 100 on the SR560. This comes up to an extra gain of 46, meaning multiplying the seismometer's counts by 1/46. |
Attachment 1: m_per_m.png
|
|
6005
|
Fri Nov 25 12:46:13 2011 |
Mirko | Update | WienerFiltering | Wiener filtering tryout |
Tried the wiener filter with the TF from p.5900
Tried it out with the TFs from p.5900:

Adding a filter element that compensates the acutator TF makes the MC lose lock. |
1337
|
Wed Feb 25 11:48:02 2009 |
Jenne | Update | PEM | Wiener filtering update - work on filtering some S5 DARM_CTRL data |
Quick update on my wiener filtering status:
Joe has been helping me get on the GRID, so I now have a grid certificate, and accounts on most/all of the clusters.
Joe also helped me get menkar to get S5 data so that I can do wiener filtering to the back-data.
I've been running the wiener filtering algorithm, and right now, it doesn't do anything to improve the DARM_CTRL data. I am confident that this is because something is funky in the wiener filtering algorithm somewhere. The indicator of this is that the wiener filtering calculation takes the same amount of time (~95 seconds) to calculate a filter for 64 seconds of data as for 1 hour of data (both for N = 2000 taps).
For reference, attached are my plots for the wiener filtering result for (1) 64 seconds of S5 data, and for (2) 3600 seconds of S5 data.
These plots were made using H1:DARM_CTRL as the signal to minimize, with 4 seismometers as the witness channels (EX_SEISX, EY_SEISY, LVEA_SEISX, LVEA_SEISY)
I'm working on figuring out what's going on with the filtering algorithm, and why it does work for C1:MC_L minimization, but does not work for H1:DARM_CTRL minimization.
|
Attachment 1: h1_DARM_64s_4seis_25Feb09.png
|
|
Attachment 2: h1_DARM_3600s_4seis_25Feb09.png
|
|
11366
|
Fri Jun 19 16:54:20 2015 |
Jenne | Update | Computer Scripts / Programs | Wiener scripts in scripts directory |
I have put the Wiener filter scripts into /opt/rtcds/caltech/c1/scripts/Wiener/ . They are under version control.
The idea is that you should copy ParameterFile_Example.m into your own directory, and modify parameters at the top of the file, and then when you run that script, it will output fitted filters ready to go into Foton. (Obviously you must check before actually implementing them that you're happy with the efficacy and fits of the filters).
Things to be edited in the ParameterFile include:
- Channel names for the witness sensors (which should each have a corresponding .txt file with the raw data)
- Channel name for the target
- Folder where this raw data is saved
- Folder to save results
- 1 or 0 to determine if need to load and downsample the raw data, or if can use pre-downsampled data
- This should probably be changed to just look to see if the pre-downsampled data already exists, and if not, do the downsampling
- 1 or 0 to determine if should use actuator pre-weighting
- Data folder for measured actuator TFs (only if using actuator pre-weighting)
- Actuator TFs can be many different exported text files from DTT, and they will be stitched together to make one set of measurements, where all points have coherence above some quantity (that you set in the ParameterFile)
- Coherence threshold for actuator data (only use data points with coherence above this amount)
- Fit order for actuator transfer function's vectfit
- 1 or 0 to decide if should use preweighting filter
- zeros and poles for preweighting filters
- 1 or 0 to decide if should use lowpass after Wiener filters (will be provided corresponding SOS coefficients for this filter, if you say yes)
- Lowpass filter parameters: cuttoff freq, order and ripple for the Cheby filter
- New sample rate for the data
- Number of Wiener filter taps
- Decide if use brute force matrix inversion or Levinson method
- Calibrations for witnesses and target
- Fit order for each of the Wiener filters
I think that's everything that is required.
|
4968
|
Thu Jul 14 17:34:35 2011 |
Ishwita, Manuel | HowTo | WienerFiltering | Wiener-Hopf equations |
Since we are using Wiener filtering in our project, we studied the derivation of Wiener-Hopf equations. Whatever we understood we have written it as a pdf document which is attached below... |
Attachment 1: derivwf.pdf
|
|
16346
|
Mon Sep 20 15:23:08 2021 |
Yehonathan | Update | Computers | Wifi internet fixed |
Over the weekend and today, the wifi was acting bad with frequent disconnections and no internet access. I tried to log into the web interface of the ASUS wifi but with no success.
I pushed the reset button for several seconds to restore factory settings. After that, I was able to log in. I did the automatic setup and defined the wifi passwords to be what they used to be.
Internet access was restored. I also unplugged and plugged back all the wifi extenders in the lab and moved the extender from the vertex inner wall to the outer wall of the lab close to the 1X3.
Now, there seems to be wifi reception both in X and Y arms (according to my android phone).
|
16350
|
Mon Sep 20 21:56:07 2021 |
Koji | Update | Computers | Wifi internet fixed |
Ug, factory resets... Caltech IMSS announced that there was an intermittent network service due to maintenance between Sept 19 and 20. And there seemed some aftermath of it. Check out "Caltech IMSS"
|
2849
|
Tue Apr 27 11:16:13 2010 |
josephb | Configuration | CDS | Wiki page with CDS .mdl names, shared memory allocation |
I've added a new page in the wiki which describes the current naming scheme for the .mdl model files used for the real time code generator. Note, that these model names do not necessarily have to be the names of the channels contained within. Its still possible to make all suspension related channels start with C1:SUS- for example. I'm also allocating 1024 8 byte channels for shared memory address space for each controller and each simulated plant.
The wiki page is here
Name suggestions, other front end models that are needed long term (HEPI is listed for example, even though we don't have it here, since in the long run we'd like to port the simulated plant work to the sites) are all welcome. |
11345
|
Wed Jun 3 17:04:08 2015 |
ericq | Update | PEM | Wilcoxon Accelerometer Huddle Test |
I've set up the Wilcoxon accelerometers on the SP table for a huddle test, to estimate their noise levels.
They're clamped down to the table along the same axis, and their housings are in good contact, in hopes to make them as correlated as possible.

Steve helped me move the DAQ cables from under the BS/PRM oplev table, to dropping from the cable tray above the POX table, so I could get them plugged in at the SP table. This also helps reduce the rats nest by the access connector, and is a fine location for when the accelerometers are attached at MC1 & MC2.
A quick glance at DTT shows coherence of >0.9 from about 2-100Hz for all six. A three-corner-hat deal will provide an easy estimate of the noise floor of each one. If we feel like being fancy / accounting for possible gain differences, we could try some MISO wiener action, but that is probably overkill. |
Attachment 1: AccHuddle.jpg
|
|
11350
|
Wed Jun 10 02:50:39 2015 |
ericq | Update | PEM | Wilcoxon Accelerometer Huddle Test |
Here are some results for the 3-corner hat subtraction for the six accelerometers, from ~1 hour of data that didn't look to have any sharp features/glitches from human activity in the lab.
I used the python uncertainties package to try and get an estimate of the uncertainty in the subtracted noise floor, by taking into account every possible possible combination of 3 sensors and the fluctuations in the spectrograms of the subtracted signals. I've attached the python code if anyone is interested in the implementation.
I pulled out the accelerometer data sheets to read their slightly varying V/g calibration (which differs on the order of a few percent from unit to unit). This improved the subtraction factor from ~20 to over 100 at some frequencies. I've edited the filter modules such that the OUT_DQ channels are already calibrated into m/s^2.

|
Attachment 1: hats2Acc.png
|
|
Attachment 2: 3hatcode.zip
|
11367
|
Sun Jun 21 13:21:03 2015 |
rana | Update | PEM | Wilcoxon Accelerometer Huddle Test |
To improve our sensor noise estimate, the ACC cables should be clamped using a rubber pad and a big dog clamp on the SP tabletop before exiting the table. Also the sensors themselves should be covered with a foam box or a double cardboard box. The high-frequency acoustic noise may limit the 10 Hz performance since piezos are not very linear.
Quote: |
I've set up the Wilcoxen accelerometers on the SP table for a huddle test, to estimate their noise levels.
They're clamped down to the table along the same axis, and their housings are in good contact, in hopes to make them as correlated as possible.
|
Wilcoxon. Not Wilconox or Wilco or Wilcoxen. Have some pity on future keyword searchers. |
11391
|
Sun Jul 5 18:14:13 2015 |
Ignacio | Update | PEM | Wilcoxon Accelerometer Huddle Test |
Updated: On Thursday/Friday (sorry for late elog) I was messing with Eric's Wilcoxon 731A accelerometer huddle test data that was taken without the box and cables being adjusted properly. Anyways, I performed the three cornered hat analysis as he had done but I also performed a six cornered hat method as well instead of permuting around in pairs of three accelerometers. The following plots of the ASD's show the results,

It is interesting to note the improvement at low frequencies when six accelerometers are used instead of six while at higher frequencies we can clearly see how the results are worst than the three hat results.

I decided to take a mean of all six accelerometers measured ground signal as well as that for their computed selfnoises, this is plotted below,

Notice the obvious improvement along the entire frequency band of the measurements when all accelerometers are used in the data analysis.
I also performed some Wiener filtering of this data. There was an obvious improvement in the results,
The mean of the signals is also plotted below, just as I did with the cornered hat methods,

I also compared the mean self noise of the accelerometers against the manufacturers calculated selfnoise that Rana put up on Github. Both methods are compared against what the manufacturer claims,

As expected the measured noise curves of the Wilcoxon is not as good as what the manufactures stated. This should improve once we redo the huddle test with a better experimental setup. We have some pretty interesting results with the six cornered hat method at around 5 Hz, it is surprisingly accurate given how rough the calculations seemed to be.
I have attached my code for reference: code_accel.zipselfnoise_allsix.png
SEE attachments for better plots of the six accelerometers... |
Attachment 5: code_accel.zip
|
Attachment 6: selfnoise_allsix_miso.png
|
|
Attachment 8: selfnoise_allsix.png
|
|
9013
|
Thu Aug 15 09:34:12 2013 |
Steve | Update | General | Wilcoxon cables rescued |
Eric and Steve,
We removed Wilcoxon Accelerometer PS and Amplifier unit under the BS optical tabel yesterday. The six cabels going to DAQ were labeled and left in place. Gain setting were 100, except channel 3 was 10.
The ~ 40 m long 2 sets of 3 cables were very happy to get their kinks out. Especially the set going just south of ITMX optical table.
We have to take better care of these cables! Your data will be useless this way. |
Attachment 1: rescuedGraycables.jpg
|
|
Attachment 2: wilconoxOut.jpg
|
|
Attachment 3: chanGains.jpg
|
|
2560
|
Tue Feb 2 15:30:03 2010 |
Alberto | Frogs | Treasure | Wild Oats |
FYI. Sitting on the top shelf of George I found an opened jar of raspberry jam and an opened jar of creamy peanut butter. Both are branded Wild Oats Market.
Wikipedia:
"Wild Oats Markets was an operator of natural foods stores and farmers markets in North America... Whole Foods officially completed their buyout of Wild Oats on August 27, 2007 [...]" |
2373
|
Wed Dec 9 18:01:06 2009 |
Koji | Update | COC | Wiping finished |
[Kiwamu, Jenne, Alberto, Steve, Bob, Koji]
We finished wiping of four test masses without any trouble. ITMY looked little bit dusty, but not as much as ITMX did.
We confirmed the surface of the ITMX again as we worked at vertex a lot today. It still looked clean.
We closed the light doors. The suspensions are left free tonight in order to check their behavior.
Tomorrow morning from 9AM, we will replace the door to the heavy ones. |
1101
|
Thu Oct 30 11:07:25 2008 |
Yoichi | Update | Computers | Wireless bridges arrived |
Five wireless bridges for the GPIB-Ethernet converters arrived.
One of them had a broken AC adapter. We have to send it back.
I configured the rest of the bridges for the 40MARS wireless network.
One of them was installed to the SR785.
I put the remaining ones in the top drawer of the cabinet, on which the label printers are sitting.
You can use those to connect any network device with a LAN port to the 40MARS network. |
2347
|
Mon Nov 30 11:45:54 2009 |
Jenne | Update | Computers | Wireless is back |
When Alberto was parting the Red Sea this morning, and turning it green, he noticed that the wireless had gone sketchy.
When I checked it out, the ethernet light was definitely blinking, indicating that it was getting signal. So this was not the usual case of bad cable/connector which is a known problem for our wireless (one of these days we should probably relay that ethernet cable....but not today). After power cycling and replugging the ethernet cable, the light for the 2.4GHz wireless was blinking, but the 5GHz wasn't. Since the wireless still wasn't working, I checked the advanced configuration settings, as described by Yoichi's wiki page: 40m Network Page
The settings had the 5GHz disabled, while Yoichi's screenshots of his settings showed it enabled. Immediately after enabling the 5GHz, I was able to use the laptop at Alberto's length measurement setup to get online. I don't know how the 5GHz got disabled, unless that happened during the power cycle (which I doubt, since no other settings were lost), but it's all better now.
|
1668
|
Thu Jun 11 14:54:18 2009 |
josephb, alberto | Update | Computers | Wireless network |
After poking around for a few minutes several facts became clear:
1) At least one GPIB interface has a hard ethernet connection (and does not currently go through the wireless).
2) The wireless on the laptop works fine, since it can connect to the router.
3) The rest of the martian network cannot talk to the router.
This led to me replugging the ethernet cord back into the wireless router, which at some point in the past had been unplugged. The computers now seem to be happy and can talk to each other.
|
6463
|
Wed Mar 28 21:15:53 2012 |
rana | Omnistructure | Computers | Wireless router for GC |
I installed a NETGEAR Wireless Router (WPN824N) today on the 131 network. The admin password for it as well as the wireless access password are in the usual places.
The SSID is 40EARTH. I have set it to allow WPA as well as WPA2 access, so the speed is only 54 Mbps for now. In a year or so, we can turn off the WPA support and up the speed. |
6465
|
Thu Mar 29 13:23:05 2012 |
Jenne | Omnistructure | Computers | Wireless router for GC |
Quote: |
I installed a NETGEAR Wireless Router (WPN824N) today on the 131 network. The admin password for it as well as the wireless access password are in the usual places.
The SSID is 40EARTH. I have set it to allow WPA as well as WPA2 access, so the speed is only 54 Mbps for now. In a year or so, we can turn off the WPA support and up the speed.
|
This router was confiscated by the GC guys this morning around ~10am. They barged in and said that someone at the 40m had connected a new router, and we had magically taken down half of the GC network. The cable was plugged in to the wrong port on the back of the router.
Junaid / Christian said that they would "secure" the router, and then reinstall it. Apparently just having a password didn't satisfy them. This was the compromise, versus them just taking the router and never bringing it back.
|
Attachment 1: IMG_0079.JPG
|
|
6467
|
Thu Mar 29 19:13:56 2012 |
Jamie | Omnistructure | Computers | Wireless router for GC |
I retrieved the newly "secured" router from Junaid. It had apparently been hooked up to the GC network via it's LAN port, but it's LAN services had no been shut off. It was therefore offering a competing DHCP server, which will totally hose a network. A definite NONO.
The new SSID is "40mWiFi", it's WPA2, and the password is pasted to the bottom of the unit (the unit is back in it's original spot on the office computer rack. |
12649
|
Wed Nov 30 11:56:56 2016 |
Lydia | Update | CDS | Wiring for Acromag auxex replacement |
I've attached a schematic for how we will connect the Acromag mosules to the slow channel I/O curently going to c1auxex. The following changes are made:
- We are getting rid of the slow readbacks from the Anti-Image and Oplev boards, as Rana says they are unnnecessary.
- The whitening switching for the QPD is currently done by a Contec "fast" binary I/O module, but can be managed by acromag instead. This alllows CAB_1Y9_34 to be fed directly into the Acromag box since all of its connections can now be managed slow.
- There's no need to change the PD whitening scheme around (since the signals never get huge), so we can set those to always be on and then lose those Contec channels. This means all of the necessary pins on CAB_1Y9_10 can go to Acromag.
- All the other backplane cables go the the fast machines only.
|
Attachment 1: auxex_acromag.pdf
|
|
14287
|
Fri Nov 9 22:24:22 2018 |
Jon | Omnistructure | | Wiring of Vacuum Acromag Chassis Complete |
Wiring of the power, Ethernet, and indicator lights for the vacuum Acromag chassis is complete. Even though this crate will only use +24V DC, I wired the +/-15V connector and indicator lights as well to conform to the LIGO standard. There was no wiring diagram available, so I had to reverse-engineer the wiring from the partially complete c1susaux crate. Attached is a diagram for future use. The crate is ready to begin software developing on Monday. |
Attachment 1: IMG_2987.jpg
|
|
Attachment 2: IMG_2985.jpg
|
|
Attachment 3: IMG_2986.jpg
|
|
4839
|
Mon Jun 20 11:04:03 2011 |
Nicole | Update | SUS | Work Plan for Week 2 |
Here is my work plan for this week:
Current Week Plan (Week 2) (As of 6/17/11)
Setting Up for Horizontal Displacement Measurements
1) Help Steve clean small table for experiment
2) Remove aluminum base from TT suspension
3) Mount shaker onto table base
4) Mount horizontal slider onto table base
5) Connect TT suspension, shaker, and horizontal slider
Begin Assembly of Sensors
1) Begin building circuit for displacement photosensors
2) Calibrate photosensor using linear regions of power versus distance curves
3) Circuit box for photosensors?
|
4416
|
Fri Mar 18 17:55:58 2011 |
Suresh | Configuration | Green Locking | Work Plan for Y-end Aux laser installation |
A rough time-table and the various tasks are given below:
Note: 700mW NPRO sitting on AP table (Model No: 126-1064-700, Sl No. 415) = Alberto's laser
Y-arm Aux laser installation
1 |
Temperature dependence of frequency of Alberto's laser:
a) Shifting Alberto's Laser (AL) to the PSL table and setting up a beat frequency measurement between AL and PSL
b) Determining the frequency vs Temperature curve for the AL
|
Mar 21st to 25th |
Bryan and Suresh |
2 |
Re-positioning the Input beam onto the IP-ANG-PD and realigning the X-arm |
Mar 21st to 25th |
Kiwamu and his 'team' :-)
|
3 |
Repositioning the optics on the Y-end table and relocating Alberto's laser ( at this point it will be rechiristened as Y-End-NPRO )
|
Mar 27th - 28th
|
Bryan and Suresh |
4 |
Maximising the doubling effiiciency and obtaining the PD and QPD signals into the CDS |
Mar 29th - Apr 1st |
" |
5 |
Aligning the Y-end green to pass through the Y-arm and locking the green to the Y arm |
Apr 3 - 8th |
" |
6 |
Aligning the IR beam to the Y- arm and locking the Y arm to the IR |
Apr 10 - 15 |
" |
|
3213
|
Wed Jul 14 10:00:14 2010 |
josephb | Update | Phase Camera | Work near 1Y2 yesterday |
Razib and I were attempting to get the output of a photodiode (PD55A in this case) recorded, so that we could independently measure the slow (~1-10 Hz) fluctuations of the light incident on the camera. This would then allow us to subtract those fluctuations out, letting us get at the camera noise in the case with signal present (as opposed to just a dark noise measurement when we look at the noise with no signal present).
Originally I was thinking of using one empty patch panel BNCs used for PEM channels down by the 1Y7 rack and go through a 110B, although Alberto pointed out he had recently removed some monitoring equipment, which watched the amplitude modulation at various frequencies of the RF distribution (i.e. 33 MHz, etc). This equipment output a DC voltage proportional to the amplitude of the RF signals. The associated channel names were C1:IOO-RFAMPD_33MHZ, C1:IOO-RFAMPD_33MHZ_CAL, C1:IOO-RFAMPD_133MHZ, etc. These are slow channels, so I presume they enter in via the slow computers, probably via pentek (I didn't check that, although in hindsight I probably should have taken the time to find exactly where they enter the system). The connections them selves were a set of BNCs on the south side, half way up the 1Y2 rack.
We simply chose one, the 33 MHz channel in this case, and connected. At around this time, the MC did become unlocked, although it looked like it was due to the MC2 watchdog tripping. The initial theory was we had bumped the Mode Cleaner while looking around for some BNC cables, although from what Rana had to do last night, it probably was the connection. We were able to restore the watchdog and confirm that the optic started to settle down again. Unfortunately, I had to leave about 5 minutes later, and didn't do as thorough an investigation as was warranted. |