ID |
Date |
Author |
Type |
Category |
Subject |
7714
|
Thu Nov 15 02:18:24 2012 |
Den | Update | Modern Control | BS oplev | I've applied LQR feedback technique to BS oplev in pitch. I think the most inconvenient thing in using LQR controller is the amount of additional states created during cost function shaping. It requires 1 filter bank for each state. To avoid this I wrote state estimation code so all states are calculated inside one function.
On the plots below cost function and oplev feedback controller performance are shown.

|
8048
|
Fri Feb 8 23:22:48 2013 |
Den | Summary | Modern Control | progress report | I wrote a small document on the application of LQG method to a Fabry-Perot cavity control. |
10744
|
Mon Dec 1 20:12:52 2014 |
ericq | Update | Modern Control | Rebirth of CESAR / ESCOBAR | I've uploaded a note at T1400735 about a new implementation of CESAR ESCOBAR ideas I've been working on. Please send me any and all feedback, comments, criticisms!
Using the things I talk about in there, I was able to have a time domain simulation of a 40m arm cavity transition through three error signals, without hardcoding the gains, offsets, or thresholds for using the signals. Some results look like this:

I'm going to be trying this out on the real IFO soon... |
10941
|
Mon Jan 26 21:10:04 2015 |
Jenne | Update | Modern Control | Waking up the OAF | I had a look at the OAF model today.
Somehow, the screens that we had weren't matching up with the model. It was as if the screens were a few versions old. Anyhow, I found the correct screens in /userapps/oaf/common/medm, and copied them into the proper place for us, /userapps/isc/c1/medm/c1oaf. Now the screens seem all good.
I also added 2 PCIE links between the OAF and the SUS models. I want to be able to send signals to the PRM's pitch and yaw. I compiled and restarted both the oaf model and the sus model.
The OAF model isn't running right now (it's got the NO SYNC error), but since it's not something that we need for tonight, I'll fix it in the morning.
My thought for trying out the OAF is to look at the coherence between seismic motion and the POP DC QPD when the PRMI is locked (no arms). I assume that the PRM is already handled in terms of angular damping (local and oplev), so the motion will be primarily from the folding mirrors. Then, if I can feedforward the seismometer signal to the PRM to compensate for the folding mirrors' motion, I can use the DC QPD as a monitor to make sure it's working when we're PRMI-only locked, or at low recycling gain with the arms. But, since I'm not actually using the QPD signal, this will be independent of the arm power increase, so should just keep working.
Anyhow, that's what my game plan is tomorrow for FF. Right now the T-240 is settling out from its move today, and the auto-zero after the move. |
10943
|
Tue Jan 27 14:00:05 2015 |
Jenne | Update | Modern Control | PRMI locked | PRMI is locked on sideband, starting ~4 minutes ago, to collect ASC/seismic data for feedforward.
|
10959
|
Fri Jan 30 02:57:03 2015 |
Jenne | Update | Modern Control | First try with PRCL ASC Wiener filtering | [Aside - How do you rotate plots in the new elog? It's showing them correctly in the attachments list below the entry, but not in the body of the log :( ]
I tried a round of PRCL ASC Wiener filtering today, but something wasn't right. I was able to either make the cavity motion worse, or completely throw the cavity out of lock. Making it less noisy didn't happen.
I took only 9 minutes of data the other day, since the PRMI didn't want to stay locked while it was daytime. So, this wasn't a whole lot to train on. But, even so, I designed some Wiener filters. The plots with the designs show the calculated Wiener filter ("Wiener") and the result from vectfit ("Fit"). Below the bode plot is the coherence between that witness (seismometer direction) and the degree of freedom (QPD channel). The fits were weighted by the choherence, so you can see that in the areas where the coherence was good, the fit was good. Elsewhere, it's not so great.






Using these filters, and assuming a Cheby1 2nd order lowpass filter at 30Hz, I predicted the following residuals:


After discovering that these filters didn't work, I went rogue and also put in a high pass filter at 0.1 Hz, but that didn't really change anything.
Here is a plot of what happened in Yaw. The Wiener filters' gains were all 0.3 here, which made the cavity motion larger, but not so large that it lost lock. The filters ought to have gains of 1 - the Wiener calculation should figure out the gains appropriately, if I've given it enough information. Here, as in the prediction plots above, red is the reference with the Wiener off, and black is with the Wiener filters on. Black is supposed to be below red, if the filters are working. Blue is the estimate of the angular motion that is being fed forward to the PRM, and you can see that at least the general shape is correct. I do need to figure out what the resonance in the blue trace is from - it's at the same frequency as a peak in the T-240's spectrum (that I didn't save). I suspect the cable might be touching the spaghetti pot on the inside, and making a mechanical short to pot vibrations.

Anyhow, more work to be done. I left the PRMI locked for a while this afternoon, starting at 5:15ish, so I'll see tomorrow how long of a lock stretch I was able to capture for training.
|
10967
|
Tue Feb 3 04:07:49 2015 |
Jenne | Update | Modern Control | Measured PRM -> POP QPD TFs | Today (after centering the POP QPD), I measured the PRM to POP QPD transfer functions. I am suspicious that this was part of my problem last week. Since most of the angular noise is coming from the folding mirrors, but I can't actuate on them, I need to know (rather, the Wiener calculator needs to know) how actuating on the PRM will affect the spot at the POP QPD.
For the plots below, I have cut out any data points that have coherence less than 0.95. I may want to go back and fill in a little bit some of the areas (particularly around 3Hz) that I had trouble getting coherence in.
Using these to prefilter my witness data, I am failing to calculate a Wiener filter. I have tried the Levinson algorithm, as well as brute-forcing it, but I'm too close to singular for either to work. I am able to calculate a set of Wiener filters without the prefiltering, or with a dummy very simple prefilter, so it's not inherently in the calculators. Separately, I can plot my vectfit-ed actuator TFs, and I can convert them to a discrete fiilter with the bilinear transform, and then use sosfilt to filter some white noise data, which comes out with the shape I expect. So. It's not inherently the filters either. More work to do, when it's not 4am.
Here are the measured actuator transfer functions. They were measured as usual with DTT, but since the measurement kept getting interrupted (MC unlock, or I wanted to add more integrations or more cycles), these are several different DTT files stitched together. In both cases I am acuating PRM's ASC[pit, yaw] EXC point, and looking at the POP QPD.


|
10993
|
Tue Feb 10 02:55:29 2015 |
Jenne | Configuration | Modern Control | Quacky filters | I discovered that somehow my Wiener filters that show up in Foton are not the same as what I have in Matlab-land.
I have plotted each of my 3 filters that I'm working with right now (T-240 X, Y and Z for PRCL Pitch) at several stages in the filter creation process. Each plot has:
- Blue trace is the Wiener filter that I want, after the vectfit process.
- Green trace is the frequency response of the SOS filters that are created by autoquack (really, quack_to_rule_them_all, which is called by autoquack). The only thing that happens in matlab after this is formatting the coefficients into a string for writing to foton.
- Red trace is the exported text file from foton.
It's not just a DC gain issue - there's a frequency dependent difference between these filters. Why??
It's not frequency warping from changing between analog and digital filters. The sample rate for the OAF model is 2048Hz, so the effect is small down at low frequencies. Also, the green trace is already discretized, so if it were freq warping, we'd see it in the green as well as red, which clearly we don't.
Has anyone seen this before? I'm emailing my seismic friends who deal in quack more than I do (BLantz and JKissel, in particular) to see if they have any thoughts.
Also, while I'm asking questions, can autoquack clear filters? Right now I'm overwriting old filters with zpk([],[],1)'s, which isn't quite the same. (I need this because the Wiener code needs more than one filter module to fit all of the poles I need, and it decides for itself how many FMs it needs by comparing the length of the poles vector with 20. If one iteration needs 4 filter modules, but the next iteration only wants 3, I don't want to leave in a bogus 4th filter.
Here are the plots:



(The giant peak at ~35Hz in the Z-axis fiilter is what tipped me off that something funny was going on) |
11186
|
Tue Mar 31 22:27:43 2015 |
Jenne | Update | Modern Control | Preliminary PRMI angular Wiener results | Before locking for the evening, I wanted to try again implementing the Wiener filters that I had designed back in Jaunary (elog 10959).
The problem then was that the newer version of Quack that I was using was doing weird things to me (elog 10993). But, tonight I used the old quack3andahalf that we used to use for Wiener-related things, and that worked (for up to order 20 filters). Actually, the pitch z-axis Wiener filter, when I copy the command string into Foton, says "Error" in the alternate box (the lower one). I also get this error message if I try to put in filters that were greater than order 20, and have been split into several filters. I'm not sure what's wrong, so for tonight I'm leaving out the pitch z-axis seismometer feed forward, and only using 20th order filters for all the rest.
So, pitch has feed forward signals from the T-240's x and y axes, and yaw has feed forward signals from all 3 seismometer channels.
At first, I just had the calculated Wiener filters, and a 10Hz lowpass, but the POP beam spot on the camera was getting slowly pushed away from the starting location. So, I added a 0.01Hz cheby1 highpass filter, and that seems to have fixed that problem. I need to go back to the simulations though, and see if this is going to cause extra noise to be injected (because of incorrect phase in the feed forward signal) at very low frequencies. All 5 Wiener filter banks have a gain of -1.
I'm getting a factor of 4-5ish between 2Hz and 3Hz in both pitch and yaw. What's interesting is that despite no direct angular suppression (as measured by the QPD) at higher frequencies, both POP22 and POPDC see improvement over a much broader range of frequencies. I'll have to think about how to predict this RIN coupling in my budgets.
The time series data for these filters was collected 2 months ago, on the 29th of January. So, it's nice to see that they work now too (although we have already seen that length feed forward signals are good over a many-month period).
In uncalibrated units (I need to calibrate the QPD to microrad, and should probably quote the PD signals in RIN), here is the plot. Blue trace (taken first) was with the feed forward on. Red trace (taken immediately afterward) was with feed forward off. This data is all PRMI-only, locked on REFL165 using Koji's recipe from elog 11174, including changing REFL165 phase to -14deg (from the -110 I found it at) for the no-arms case.
PRMI_31Mar2015.pdf |
11202
|
Sat Apr 4 17:54:03 2015 |
rana | Update | Modern Control | Preliminary PRMI angular Wiener results | This is a very cool result. I'm surprised that it can work so good. Please post what frequency dependent weighting you used on the target before running the Wiener code.
I think its important to tune it to keep the low frequencies from getting amplified by a factor of 10 (as they are in your plots). The seismometers are all just noise acting like thermometers and tilt sensors below 0.1 Hz. Temperature and tilt do not couple to our interferometer very much.
It also seems weird that you would need 20th order filters to make it work good. In any case, you can always split the SOS up into pieces before making the digital filters. For LLO, we used 3 buttons in some cases.
Quote: |
Before locking for the evening, I wanted to try again implementing the Wiener filters that I had designed back in Jaunary (elog 10959).
|
|
11229
|
Tue Apr 21 01:17:13 2015 |
Jenne | Update | Modern Control | T-240 self-noise propagated through stack and pendulum | Going back to Wiener filtering for a moment, I took a look at what the T-240 noise level looks like in terms of pitch motion on one of our SOS optics (eg. PRM).
The self-noise of the T-240 (PSD, in dB referenced to 1m^2/s^4/Hz) was taken by pulling numbers from the Users Guide. This is the ideal noise floor, if our installation was perfect. I'm not sure where Kissel got the numbers from, but on page 13 of G1200556 he shows higher "measured" noise values for a T-240, although his numbers are already transformed to m/rtHz.
To get the noise numbers to meters, I use: . The top of that fraction is (a) getting to magnitude from power-dB and (b) getting to asd units from psd units. The bottom of the fraction is getting rid of the extra 1/s^2.
Next I propagate this seismometer noise (in units of m/rtHz) to effective pendulum pitch motion, by propagating through the stacks and the transfer function for pos motion at the anchor point of the pendulum to pitch motion of the mirror (see eq 63 of T000134 for the calculation of this TF). This gives me radians/rtHz of mirror motion, caused by the ground motion:
.
I have not actually calibrated the POP QPD, so I will need to do that in order to compare this seismometer noise to my Wiener filter results.
|
11310
|
Tue May 19 14:51:44 2015 |
ericq | Update | Modern Control | Brushing up on Wiener Filtering | As part of preparing for the SURF projects this summer, I grabbed ~50 minutes of MCL and STS_1 data from early this morning to do a little MISO wiener filtering. It was pretty straightforward to use the misofw.m code to achieve an offline subtraction factor of ~10 from 1-3Hz. This isn't the best ever, but doesn't compare so unfavorably to older work, especially given that I did no prefiltering, and didn't use all that long of a data stretch.
Code and plot (but not data) is attached. |
11313
|
Tue May 19 17:53:38 2015 |
rana | Update | Modern Control | Brushing up on Wiener Filtering | Good to see that misofw.m is still alive and well. Todo:
- Apply weighting to the filter via time domain SOS prefiltering of the signals (see Jenne's code from the LLO Global FF paper)
- Consider using some of the MC SUSPOS signals. Its not strictly legal, but I wonder if its responsible for out noise below 1 Hz.
- Steve was in the midst of hooking up the Wilcoxones to get better subtraction at 5-15 Hz, but couldn't find cables. We need to hunt them down in W Bridge.
- Attach one of the medium or big Lings to the MC2 chamber platform and see if we can do this in hardware without driving the suspension. WE want to use a DAC channel (from where?) and pipe it to the Ling using a medium high current drive. Might start with the 50 Ohm output of the SR560 and later use a BUF634 ala Crackle coil driver.
|
17535
|
Tue Apr 4 11:10:19 2023 |
Anchal | Summary | NeuralNet | Testing neural network controller during day time | I ran two recently trained neural network controllers today between 10 am and Noon. Each test comprised of four segments:
- All loops open
- Linear loop closed
- Neural Network working alone
- Neural Network + Linear Loop
The latest controller unfortunately failed in both cases, working alone and working together with linear loop.
The second latest controller functioned well, keeping the arm locked throughout. |
17537
|
Thu Apr 6 11:49:45 2023 |
Anchal | Summary | NeuralNet | Testing neural network controller during day time | I've turned on NN controller for MC WFS PIT loops. One can disable this controller and go back to linear controller by sitemap>IOO>C1IOO_WFS_MASTER>Actions!>Stop Shimmer . One can start the controller again sitemap>IOO>C1IOO_WFS_MASTER>Actions!>Run Shimmer . |
17538
|
Thu Apr 6 21:09:12 2023 |
Koji | Summary | NeuralNet | Testing neural network controller during day time | I'm going to get into the PSL enclosure. Also turn on the HEPA for a while during the intrusion.
Work done.
Start of the work:
(cds) ~>gpstime
PDT: 2023-04-06 21:09:36.670623 PDT
UTC: 2023-04-07 04:09:36.670623 UTC
GPS: 1364875794.670623
End of the work:
(cds) ~>gpstime
PDT: 2023-04-06 21:19:37.331716 PDT
UTC: 2023-04-07 04:19:37.331716 UTC
GPS: 1364876395.331716 |
17548
|
Wed Apr 19 09:52:50 2023 |
Radhika | Update | NeuralNet | Rayleigh spectrograms | Attached are the Rayleigh spectrograms of the error/control signal channels associated with the NN nonlinear control of IMC (pitch). The 4-hour data stretch starts at 3:45pm PDT on 4/18. The spectrograms were generated with (stride=5, fftlength=2, overlap=1). PNG images are attached for reference; the generated pdf files were too large to include here or send over email.
The Rayleigh statistic measures nongaussianity of the data. |
14603
|
Fri May 10 18:24:29 2019 |
gautam | Update | NoiseBudget | aligoNB | I pulled the aligoNB git repo to /ligo/GIT/aligoNB/aligoNB. There isn't a reqs.txt file in the repo so installing the dependencies on individual workstations to get this running is a bit of a pain. I found the easiest thing to do was to setup a virtual environment for the python3 stuff, this way we can run python2 for the cdsutils package (hopefully that gets updated soon). I'm setting up a C1 directory in there, plan is to budget some subsystems like Oplev, ALS for now, and develop the code for the eventual IFO locking. As a test, I ran the H1 noise budget (./aligonb H1), works, so looks like I got all the dependencies... |
15047
|
Mon Nov 25 22:10:26 2019 |
shruti | Update | NoiseBudget | Diagnostics | This is to help troubleshoot the excess noise measured earlier.
The following channels were measured at GPS times 1258586880 s and 1258597457 s, corresponding to low and high Power Recycling Gain (PRG) respectively.
Excess noise was seen between 25-110 Hz in the high PRG case when compared to the low PRG case in the following channels:
C1:LSC-CARM-IN1_DQ (shown in Attachment 1 where the reference is low PRG)
C1:ALS-Y_ERR_MON_OUT_DQ
C1:ALS-BEAT{X,Y}_FINE_PHASE_OUT_DQ
C1:SUS-ETM{X,Y} _SENSOR_{LL,LR,UL,UR}
C1:ALS-TRX_OUT_DQ
Surprisingly, it was also seen to a smaller extent in (refer Attachment 3)
C1:SUS-ITMX_SENSOR_{LL,LR,UL,UR}
A different type of noise spectrum, attributed to known electronic effects, was observed for
C1:SUS-ITMY_SENSOR_{LL,UL} (refer Attachment 2)
These did not show any significant change in the noise spectrum:
C1:LSC-DARM-IN1_DQ (shown in Attachment 1 where the reference is low PRG)
C1:ALS-X_ERR_MON_OUT_DQ
C1:ALS-TRY_OUT_DQ
C1:SUS-ITMY_SENSOR_{LL,LR,UL,UR}
C1:SUS-ITMY_SENSOR_{LR,UR} (refer Attachment 2)
Broadband noise in:
C1:LSC-PO{X,Y}11_I_ERR_DQ
|
15300
|
Tue Apr 7 15:30:40 2020 |
Jon | Summary | NoiseBudget | 40m noise budget migrated to pygwinc | In the past year, pygwinc has expanded to support not just fundamental noise calculations (e.g., quantum, thermal) but also any number of user-defined noises. These custom noise definitions can do anything, from evaluating an empirical model (e.g., electronics, suspension) to loading real noise measurements (e.g., laser AM/PM noise). Here is an example of the framework applied to H1.
Starting with the BHD review-era noises, I have set up the 40m pygwinc fork with a working noise budget which we can easily expand. Specific actions:
- Updated the 40m fork to the latest pygwinc version (while preserving the commit history).
- Added a directory
./CIT40m containing the 40m-specific noise budget files (created by GV).
- Added an ipython notebook
CIT40m.ipynb at the root level showing how to generate a noise budget.
- Integrated our DAC and seismic noise estimators into pygwinc.
- Marked the old 40m NB repo as obsolete (last commit > 2 yrs ago). Many of these noise estimates are probably stale, but I will work with GV to identify which ones can be migrated.
I set up our fork in this way to keep the 40m separate from the main pygwinc code (i.e., not added to as a built-in IFO type). With the 40m code all contained within one root-level directory (with a 40m-specific name), we should now always be able to upgrade to the latest pygwinc without creating intractable merge conflicts. |
16152
|
Fri May 21 12:12:11 2021 |
Paco | Update | NoiseBudget | AUX PDH loop identification | [Anchal, Paco]
We went into 40m to identify where XARM PDH loop control elements are. We didn't touch anything, but this is to note we went in there twice at 10 AM and 11:10 AM. |
16975
|
Wed Jul 6 19:58:16 2022 |
Paco | Summary | NoiseBudget | XARM noise budget | [Anchal, Paco, Rana]
We locked the XARM using POX11 and made a noise budget for the single arm displacement; see Attachment #1. The noise budget is rough in that we use simple calibrations to get it going; for example we calibrate the measured error point C1:LSC-XARM_IN1_DQ using the single cavity pole and some dc gain to match the UGF point. The control point C1:LSC-XARM_OUT_DQ is calibrated using the actuator gain measured recently by Yuta. We also overlay an estimate of the seismic motion using C1:PEM-SEIS_BS_X_OUT_DQ (calibrated using a few poles to account for stack and pendulum), and finally the laser frequency noise as proxied by the mode cleaner C1:IOO-MC_F_DQ.
A couple of points are taken with this noise budget, apart from it needing a better calibration;
- Overall the inferred residual displacement noise is high, even for our single arm cavity.
- By looking at the sim OLTF in foton, it seemed that the single arm cavity loop TF could easily become unstable due to some near-UGF-funkiness likely from FM3 (higher freq boost), so we disabled the automatic triggering on it; the arm stayed locked and we changed the error signal (light blue vs gold (REF1) trace)
- The arm cavity is potentially seeing too much noise from the IMC in the 1 to 30 Hz band in the form of laser frequency noise.
- Need IMC noise budget to properly debug.
- At high frequency (>UGF), there seem to be a bunch of "wiggles" which remain unidentified.
- We actually tried to investigate a bit into these features, thinking they might have something to do with misalignment, but we couldn't really find significant correlation.
RXA edit:
- we also noticed some weirdness in the calibration of MC_F v. Arm. We think MC_F should be in units of Hz, and Paco calculated the resulting motion as seen by the arm, but there was a factor of several between these two. Need to calibrate MC_F and check. In principle, MC_F will show up directly in ALS_BEATX (with the green PDH lock off), and I assume that one is accurately calibrated. Somehow we should get MC_F, XARM, and ALS_BEAT to all agree. JC is working on calibrating the Mini-Circuits frequency counter, so once that is done we will be in good shape.
- we may need to turn on some MC_L feedback for the IMC, so that the MC length follows the NPRO frequency below ~20 Hz.
- Need to estimate where the IMC WFS noise is in all of this. Does it limit the MC length stability in any frequency band? How do we determine this?
- Also, we want to redo this noise budget today, whilst using AS55 instead of POX. Please measure the Schnupp asymmetry by checking the optimum demod phase in AS55 for locking Xarm v Yarm.
|
17520
|
Thu Mar 23 17:47:53 2023 |
Paco | Update | NoiseBudget | LO phase noise budget (BH55_Q) | I drafted a calibrated LO Phase noise budget using diaggui whose template is saved under /opt/rtcds/caltech/c1/Git/40m/measurements/BHD/LO_PHASE_cal_nb.xml which includes new estimates for laser frequency and intensity noises at the LO phase when MICH is locked (whether they couple through MICH or the LO path is to be determined with noise coupling measurements in the near future, but we expect them to couple through the LO phat mostly).
Attachment #1 shows the result.
Laser Frequency Noise
To calibrate the laser frequency noise contribution, I used the LO PHASE error point away from the control bandwidth (~ 20 Hz) and the calibrated C1:IOO-MC_F control point (in Hz) which should represent the laser frequency noise above 100 Hz. and dithered MC2 at frequencies around to 130, 215, and 325 Hz to match the LO phase error point with the MC_F signal. I was expecting to use a single 0 Hz pole + gain (to get the phase equivalent of the laser frequency noise) but in the end I managed to calibrate with a single gain of 3.6e-7 rad/Hz and no pole. Since the way the laser frequency noise couples into our BHD readout may be complicated (especially when using BH55 RF sensor) I didn't think much of this for now.
Laser Intensity Noise
For the intensity noise, I followed more or less a similar prescription as for laser frequency noise. This time, I used the AOM in the PSL table to actuate on the 0th order intensity going into the interferometer. Attachments #2-3 show the connection made to the RF driver where I added a 50 mVpp sine (at an offset of 0.1 V) excitation in the AM port to inject intensity noise calibration lines at 215 and 325 Hz and matched the LO_PHASE error point with the BHDC_SUM noise spectrum. |
5
|
Fri Oct 19 16:11:38 2007 |
pkp | Other | OMC | OMC PZT response | Sam and I locked the laser to the OMC cavity and looked at the error signal as a function of the voltage applied to the OMC PZT.
Here are two plots showing the response as a function of frequency from 1 kHz to 100 kHz and another high-res response in the region of 4.5 kHz to 10 kHz. |
6
|
Sat Oct 20 11:54:13 2007 |
waldman | Other | OMC | OMC and OMC-SUS work | [Rich, Chub, Pinkesh, Chris, Sam]
Friday the 18th was a busy day in OMC land. Both DCPDs were mounted to the glass breadboard and the OMC-SUS structure was rebuilt to the point that an aluminum dummy mass is hanging, unbalanced. The OSEMs have not be put on the table cloth yet, but everything is hanging free. As for the DCPDs, if you recall one beam is 3mm off center from the DCPD tombstone. Fortunately, one DCPD is nearly 3mm offcenter from the case in the right direction, so the errors nearly cancel. The DCPD is too high, so the beam isn't quite centered, but they're close. We'll get photos of the beam positions in someday. Also, the DC gain between the two PDs is, at first glance, different by 15%. DCPD1, the one seen in transmission has 315 mV of signal while DCPD2 has 280 mV. Not sure why, could be because of beam alignment or tolerances in the Preamp or the angle incident on the diode or the QE of the diodes. The glass cans have *not* been removed.
|
8
|
Mon Oct 22 19:27:14 2007 |
pkp | Other | OMC | PZT calibration/ transfer function. | We measured the PZT transfer function by comparing the PZT response of the circuit with the cavity in the loop, with that of the circuit without the cavity in the loop. Basically measure the transfer function of the whole loop with the laser/PZT and Op-amps in it. Then take another measurement of the transfer function of everything else besides the PZT and from both these functions, we can calculate the PZT response.
The calibration was done by using the error signal response to a triangular wave of volts applied to the PZT. A measurement of the slope of the error signal , which has three zero-crossings as the cavity sweeps through the sidebands, gives us the Volts/Hz response. In order to derive a frequency calibration of the x axis, we assume that the first zero crossing corresponds to the first side band (-29.5 MHz) and the third one corresponds with the other sideband (+29.5 MHz). And then by using the fact that we know the response of the cavity to a constant frequency shift, we can use the Volts/Hz measurement to calculate the Volts/nm calibration. The slope that was calculated was 3.2e-6 V/Hz and using the fact that the cavity is 1 m in length and the frequency is 1064 nm, we get a calibration of 0.9022 V/nm.
|
9
|
Tue Oct 23 09:01:00 2007 |
rana | Other | OMC | PZT calibration/ transfer function. | Are you sure that the error signal sweep is not saturated on the top ends? This is usually the downfall
of this calibration method. |
14
|
Thu Oct 25 17:52:45 2007 |
waldman | Other | OMC | OMCs with QPDs | [Rich, Chub, Pinkesh, Sam]
Yesterday we got the QPD, OTAS, and PZT cabling harness integrated with the OMC. We found a few things out, not all of them good. The QPDs went on no problem and could be fairly well aligned by hand. We "aligned" them by looking at all four channels of the QPD on the scope and seeing that there is signal. Since the beam is omega = 0.5 mm, this is a reasonable adjustment. We then connected the OTAS connector to the OTAS and found that the heater on the OTAS was bonded on about 30 degrees rotated from its intended position. This rotated the connector into the beam and caused a visible amount of scattering. This wasn't really a disaster until I removed the connector from the heater and broke the heater off of the aluminum parts of the OTAS. Two steps backwards, one step forward. After the OMC, OMC-SUS integration test we will re-bond the heater to the aluminum using VacSeal. In the meantime, the OMC has been moved to Bridge 056 for integration with the OMC-SUS. More on that as we make progress. |
16
|
Thu Oct 25 23:35:36 2007 |
waldman | Other | OMC | Hang the OMC! | [Pinkesh, Sam]
We tried, convicted and hung the OMC today. The OMC was found guilty of being overweight, and unsymmetrically balanced. The unsymmetry was kind of expected and was corrected with a hefty stack of counterweights positioned over the counterweighting holes. The stacks will be measured at some future date and correctly sized objects machined. The overweightness showed up when the level hanging breadboard was about 5 mm low. This showed up in the board height above the table as well as the OSEM flag positions within their holes. The problem was remedied with a liposuction of the intermediate mass. We removed both small vertical cylinder weights that Chris added, and then we removed the heavy steel transverse weight that can be used to adjust the tip around the long axis (I forgot what its called).
The top of the breadboard ended up about 154 mm off the table. The breadboard is 39 mm thick, and the optics are centered (30 - 12.7) = 17.3 mm below the surface for a as hanging beams height of 154 -39 - 17.3 = 97.7 mm or about an 0.150 inches lower than we were aiming for. Can I get a refund?
We screwed up in multiple ways:
- The slotted disks that capture the wires do not have the alignment bore used to center the wire in the hole
- We didn't correctly route the far field QPD cable so it runs funny
- We didn't have a tool which could be used to get two of the DCPD preamp box mounting screws (which are M3's chub!)
- We don't have the cable clamps to tie off the electrical cables to the intermediate mass
- We don't have any of the cabling from the OMC-SUS top to the rack so we can't test anything
- We haven't uploaded pretty pictures for all to see
We left the OMC partially suspended by the OMC-SUS and partly resting on the installation lab jacks which are currently acting as EQ stops. After we fix the cabling we will more permanently hang it. PS, It looks like the REFL beam extraction will be tricky so we need to get on that.... |
19
|
Fri Oct 26 17:34:43 2007 |
waldman | Other | OMC | OMC + earthquake stops |
[Chub, chris, Pinkesh, Sam]
Last night we hugn the OMC for the first time and came up with a bunch of pictures and some problems. Today we address some of the problems and, of course, make new problems. We replaced the flat slotted disks with the fitted slotted disks that are made to fit into the counterbore of the breadboard. This changed the balance slightly and required a more symmetric distribution of mass. It probably did not change the total mass very much. We did find that the amount of cable hanging down strongly affected the breadboard balance and may also have contributed to the changing balance.
We also attached earthquake stops and ran into a few problems:
- The bottom plate of the EQ stops is too thick so that it bumps into the tombstones
- The vertical member on the "waist" EQ stops is too close to the breadboard, possibly interfering with the REFL beam
- The "waist" EQ stops are made from a thin plate that doesn't have enough thickness to mount helicoils in
- Helicoil weren't loaded in the correct bottom EQ stops
- The DCPD cable loops over the end EQ stop looking nasty but not actually making contact
However, with a little bit of jimmying, the EQ stops are arrayed at all points within a few mm of the breadboard. Meanwhile, Chub has cabled up all the satellite modules and DCPD modules and Pinkesh is working on getting data into the digital system so we can start playing games. Tonight, I intend to mount a laser in Rana's lab and fiber couple a beam into the 056 room so we can start testing the suspended OMC. |
20
|
Fri Oct 26 21:48:40 2007 |
waldman | Configuration | OMC | Fiber to 056 | I set up a 700 mW NPRO in Rana's lab and launched it onto a 50m fiber. I got a few mW onto the fiber, enough to see with a card before disabling the laser. The fiber now runs along the hallway and terminates in rm 056. Its taped down everywhere someone might trip on it, but don't go out of your way to trip on it or pull on it because you are curious. Tomorrow I will co-run a BNC cable and attenuate the NPRO output so it can only send a few mW and so be laser safe. Then we can try to develop a procedure to align the beam to a suspended OMC and lock our suspended cavity goodness.
Notes to self: items needed from the 40m
- ND10 and ND20 neutral density filter
- EOM and mount set for 4 inch beam height
- Post for fiber launch to get to 4 inch
- Mode matching lens at 4in
- 3x steering mirror at 4in
- RF photodiode at 4in
- Post for camera to 4in
- Light sheild for camera
- Long BNC cable
Some of these exist at 056 already |
21
|
Sat Oct 27 19:00:44 2007 |
waldman | Configuration | OMC | Hanging, locked OMC with REFL extracted. | I got the OMC locked to the fiber output today. It was much more difficult than I expected and I spent about 30 minutes or so flailing before stopping to think. The basic problem is that the initial alignment is a search in 4-dimensional space and there is naturally only one signal, the reflected DC level, to guide the alignment. I tried to eyeball the alignment using the IR card and "centering" the beams on mirrors, but I couldn't get close enough to get any light through. I also tried to put a camera on the high reflector transmission, but with 1.5 mW incident on the cavity, there is only 1.5 microwatts leaking through in the best case scenario, and much, much less during alignment.
I resolved the problem by placing a high reflector on a 3.5 inch tall fixed mount and picking off the OMC transmitted beam before it reaches the DC diodes. I took the pickoff beam to a camera. The alignment still sucked because even though the beam cleanly transmitted the output coupler, it wasn't anywhere close to getting through the OTAS. To resolve this problem, I visually looked through the back of M2 at M1 and used the IR card to align the beam to the centers of each mirror. That was close enough to get me fringes and align the camera. With the camera aligned, the rest was very easy.
I restored the PDH setup we know and love from the construction days and locked the laser to the OMC with no difficulty. The laser is in Rana's lab so I send the +/- 10V control signal from the SR560 down a cable to 058E where it goes into the Battery+resistor box, the Throlabs HV amplifier, and finally the FAST channel of the NPRO. BTW, a simple experiment sows that about 35 +/- 3 V are required to get an FSR out of the NPRO, hence the Thorlabs HV. The EOM, mixer, splitter, etc is on the edge of the table.
With this specific OMC alignment, ie. the particular sitting on EQ stops, it looks like all of the ghost beams have a good chance of coming clear. I can fit a 2 inch optic in a fixed mount in between the end of the breadboard and the leg of the support structure. A picture might or might not be included someday. One of the ghost beams craters directly into the EQ stop vertical member. The other ghost barely misses M2 on its way down the length of the board. In its current configuration, the many REFL beam misses the leg by about 1.5 inches. |
25
|
Mon Oct 29 11:07:22 2007 |
waldman | Software Installation | OMC | Software install on OMS | [Alex, Sam]
We spent a little time this morning working on OMS and getting things restarted. A few changes were made. 1) We put openmotif on OMS so that the burtrb doesn't throw that crappy libXm any more. 2) We upgraded OMS to a 32 kHz sampling rate from 2 kHz. All the filters will have to be changed. We also added a PDH filter path to maybe feedback PDH signals cuz that will be cool. Maybe someday I will write up the very cool channel adding procedure. |
26
|
Mon Oct 29 12:20:15 2007 |
waldman | Configuration | OMC | Changed OMS filters | I changed the OMS configuration so that some of the OMC-SUS LED channels go to a breakout box so that we can input the PDH error signal. After lunch, we will try to lock the cavity with a PDH error signal and digital filters. Then its on to dither locked stuff. Note that this LED business will have to be changed back some day. For now, it should be extremely visible because there are dangling cables and a hack job interface lying around. |
27
|
Mon Oct 29 23:10:05 2007 |
waldman | Configuration | OMC | Lost in DAQspace | [Pinkesh, Sam]
In setting up a Digital based control of the hanging OMC, we naively connect the Anti-Imaging filter output to an Anti-Aliasing input. This led to no end of hell. For one thing, we found the 10 kHz 3rd order butterworth at 10 kHz, where it should be based on the install hardware. One wonders in passing whether we want a 10 kHz butter instead of a 15 kHz something else, but I leave that for a later discussion. Much more bothersome is a linear phase shift between output and input that looks like ~180 microseconds. It screams "What the hell am I!?" and none of us could scream back at it with an answer. I believe this will require the Wilson House Ghost Busters to fully remedy on the morrow. |
37
|
Wed Oct 31 09:45:28 2007 |
waldman | Other | OMC | Resolution to DAQland saga | [Jay, Sam]
We did a rough accounting for the linear delay this morning and it comes out more or less correct. The 10 kHz 3rd order butterworth AA/AI filter gives ~90 degrees of phase at 6 kHz, or 42 microseconds. Taken together, the two AA and AI filters are worth 80 microseconds. The 1.5 sample digital delay is worth 1.5/32768 = 45 microseconds. The remaining 160 - 125 = 35 microseconds is most likely taken up by the 64 kHz to 32 kHz decimation routine, assuming this isn't accounted for already in the 1.5 sample digital delay.
It remains to be seen whether this phase delay is good enough to lock the laser to the OMC cavity |
42
|
Wed Oct 31 23:55:17 2007 |
waldman | Other | OMC | QPD tests | The 4 QPDs for the OMC have been installed in the 056 at the test setup. All 4 QPDs work and have medm screens located under C2TPT. The breadboard mounted QPDs are not very well centered so their signal is somewhat crappy. But all 4 QPDs definitely see plenty of light. I include light and dark spectra below. QPDs 1-2 are table-mounted and QPD 2 is labeled with a bit of blue tape. QDPs 3-4 are mounted on the OMC. QPD3 is the near field detector and QPD4 is the far field. In other words, QPD3 is closest to the input coupler and QPD4 is farthest.
Included below are some spectra of the QPDs with and without light. For QPDs 1 & 2, the light source is just room lights, while 3&4 have the laser in the nominal OMC configuration with a few mWs as source. The noise at 100 Hz is about 100 microvolts / rtHz. If I recall correctly, the QPDs have 5 kOhm transimpedance (right Rich?) so this is 20 nanoamps / rtHz of current noise at the QPD. |
43
|
Thu Nov 1 01:28:04 2007 |
waldman | Other | OMC | First digital lock of OMC | [Pinkesh, Sam]
We locked a fiber based NPRO to the suspended OMC tonight using the TPT digital control system. To control the laser frequency, we took the PZT AI output and ran it on a BNC cable down the hallway to the Thorlabs HV box. The Thorlabs is a singled ended unit so we connected the AI positive terminal only and grounded the BNC to the AI shield. We could get a -6 to 1.5 V throw in this method which fed into the 10 k resisotr + 9 V battery at the input of the HV box. The HV out ran to the NPRO PZT fast input.
We derived our error signal from a PDA255 in reflection with a 29.5 MHz PDH lock. The signal feeds into one of the unused Tip/Tilt AA channels and is passed to the PZT LSC drive through the TPT_PDH1 filter bank. In the PZT_LSC filter we put a single pole at 1 Hz which, together with the phase we mentioned the other night (180 degrees at 3 kHz) should allow a 1 kHz-ish loop. In practice, as shown below, we got a 650 Hz UGF with 45 degrees of phase margin and about 6 dB of gain margin.
The Lower figure shows the error point spectrum with 3 settings. REF0 in blue shows lots of gain peaking at 1.5 kHz-ish, just where its expected - the gain was -40. The REF1 has gain of -20 and shows no gain peaking. The current trace in red shows some gain peaking cuz the alignment is better but it also has included a 1^2:20^2 boost which totally crushes the low frequency noise. We should do a better loop sweep after getting the alignment right so we can see how much boost it will really take.
Just for fun, we are leaving it locked overnight and recording the PZT_LSC data for posterity. |
58
|
Fri Nov 2 12:18:47 2007 |
waldman | Summary | OMC | Locked OMC with DCPD | [Rich, Sam]
We locked the OMC and look at the signal on the DCPD. Plots included. |
59
|
Sat Nov 3 16:20:43 2007 |
waldman | Summary | OMC | A good day's work |
I followed up yesterday's test of the PZT with a whole mess of characterizations of the PZT control and finished the day by locking the OMC with a PZT dither lock and a 600 Hz loop. I haven't analyzed any of the data yet, so its not calibrated in physical units and etc. etc. etc. Since a lot of the sweeps below are of a "drive the PZT, look at the PDH signal" nature, a proper analysis will require taking out the loop and calibrating the signals, which alas, I haven't done. Nonetheless, I include all the plots because they are pretty. The files included below are:
- DitherLock_sweep: Sweep of the IN2/IN1 for the dither lock error point showing 600 Hz UGF
- HiResPZTDither_sweep: Sweep of the PZT dither input compared to the PDH error signal. I restarted the front end before the sweep was finished accounting for the blip.
- HiResPZTDither_sweep2: Finish of the PZT dither sweep
More will be posted later. |
60
|
Sun Nov 4 23:22:50 2007 |
waldman | Update | OMC | OMC PZT and driver response functions | I wrote a big long elog and then my browser hung up, so you get a less detailed entry. I used Pinkesh's calibration of the PZT (0.9 V/nm) to calibrate the PDH error signal, then took the following data on the PZT and PZT driver response functions.:
- FIgure 1: PZT dither path. Most of the features in this plot are understood: There is a 2kHz high pass filter in the PZT drive which is otherwise flat. The resonance features above 5 kHz are believed to be the tombstones. I don't understand the extra motion from 1-2 kHz.
- Figure 2: PZT dither path zoom in. Since I want to dither the PZT to get an error signal, it helps to know where to dither. The ADC Anti-aliasing filter is a 3rd order butterworth at 10 kHz, so I looked for nice flat places below 10 KHz and settled on 8 kHz as relatively harmless.
- Figure 3: PZT LSC path. This path has got a 1^2:10^2 de-whitening stage in the hardware which hasn't been digitally compensated for. You can see its effect between 10 and 40 Hz. The LSC path also has a 160 Hz low path which is visible causing a 1/f between 200 and 500 Hz. I have no idea what the 1 kHz resonant feature is, though I am inclined to point to the PDH loop since that is pretty close to the UGF and there is much gain peaking at that frequency.
|
63
|
Mon Nov 5 14:44:39 2007 |
waldman | Update | OMC | PZT response functions and De-whitening | The PZT has two control paths: a DC coupled path with gain of 20, range of 0 to 300 V, and a pair of 1:10 whitening filters, and an AC path capacitively coupled to the PZT via a 0.1 uF cap through a 2nd order, 2 kHz high pass filter. There are two monitors for the PZT, a DC monitor which sniffs the DC directly with a gain of 0.02 and one which sniffs the dither input with a gain of 10.
There are two plots included below. The first measures the transfer function of the AC monitor / AC drive. It shows the expected 2 kHz 2d order filter and an AC gain of 100 dB, which seems a bit high but may be because of a filter I am forgetting. The high frequency rolloff is the AA and AI filters kicking in which are 3rd order butters at 10 kHz.
The second plot is the DC path. The two traces show the transfer function of DC monitor / DC drive with and with an Anti-dewhitening filter engaged in the DC drive. I fit the antidewhite using a least squares routine in matlab constrained to match 2 poles, 2 zeros, and a delay to the measured complex filter response. The resulting filter is (1.21, 0.72) : (12.61, 8.67) and the delay was f_pi = 912 Hz. The delay is a bit lower than expected for the f_pi = 3 kHz delay of the AA, AI, decimate combination, but not totally unreasonable. Without the delay, the filter is (1.3, 0.7) : (8.2, 13.2) - basically the same - so I use the results of the fit with delay. As you can see, the response of the combined digital AntiDW, analog DW path is flat to +/- 0.3 dB and +/- 3 degrees of phase.
Note the -44 dB of DC mon / DC drive is because the DC mon is calibrated in PZT Volts so the TF is PZT Volts / DAC cts. To calculate this value: there are (20 DAC V / 65536 DAC cts)* ( 20 PZT V / 1 DAC V) = -44.2 dB. Perfect!
I measured the high frequency response of the loop DC monitor / DC drive to be flat. |
79
|
Wed Nov 7 14:01:31 2007 |
waldman | Omnistructure | OMC | Frequency and Intensity noise | One of the biggest problems I had using the PZT to lock was excessive noise. I did a little noise hunting and found that the problem was the cable running from the rack to the laser fast input. As a reminder, the laser has a 4 MHz / volt fast input. We require about 300 MHz to go one FSR, so there is a Thorlabs HV box between at the NPRO fast input which takes 0-10 V -> 0-150 V. The 150 V HV range is worth about 600 MHz of NPRO frequency.
OLD SETUP: Single side of DAC differential (10 Vpp) -> 9V in series with 10 kOhm -> 10 kOhm input impedance of Thorlabs HV -> NPRO
We used the single side of the DAC differential because we didn't have a differential receiver. This turned out to be a bad idea because the cable picks up every 60 Hz harmonic known to man kind.
NEW SETUP: Digital conditioning -> DAC differential (digitally limited to 0 - 1 V) -> SR560 in A-B mode gain 10 (0 - 10 V output)-> Thorlabs HV -> NPRO.
This has almost no 60 Hz noise and works much, much better. Moral of the story, ALWAYS USE DIFFERENTIAL SIGNALS DIFFERENTIALLY !
Note that I may be saturating the SR560 with 10 V output, Its spec'd for 10 Vpp output with 1 VDC max input. I don't know whether or not it can push 10 V out.... |
82
|
Thu Nov 8 00:55:44 2007 |
pkp | Update | OMC | Suspension tests | [Sam , Pinkesh]
We tried to measure the transfer functions of the 6 degrees of freedom in the OMS SUS. To our chagrin, we found that it was very hard to get the OSEMs to center and get a mean value of around 6000 counts. Somehow the left and top OSEMs were coupled and we tried to see if any of the OSEMs/suspension parts were touching each other. But there is still a significant coupling between the various OSEMs. In theory, the only OSEMS that are supposed to couple are [SIDE] , [LEFT, RIGHT] , [TOP1, TOP2 , TOP3] , since the motion along these 3 sets is orthogonal to the other sets. Thus an excitation along any one OSEM in a set should only couple with another OSEM in the same same set and not with the others. The graphs below were obtained by driving all the OSEMS one by one at 7 Hz and at 500 counts ( I still have to figure out how much that is in units of length). These graphs show that there is some sort of contact somewhere. I cant locate any physical contact at this point, although TOP2 is suspicious and we moved it a bit, but it seems to be hanging free now. This can also be caused by the stiff wire with the peek on it. This wire is very stiff and it can transmit motion from one degree of freedom to another quite easily. I also have a graph showing the transfer function of the longitudnal degree of freedom. I decided to do this first because it was simple and I had to only deal with SIDE, which seems to be decoupled from the other DOFs. This graph is similar to one Norna has for the longitudnal DOF transfer function, with the addition of a peak around 1.8 Hz. This I reckon could very be due to the wire, although it is hard to claim for certain. I am going to stop the measurement at this time and start a fresh high resolution spectrum and leave it running over night.
There is an extra peak in the high res spectrum that is disturbing. |
86
|
Fri Nov 9 00:01:24 2007 |
waldman | Omnistructure | OMC | OMC mechanical resonances (Tap tap tappy tap) | [Pinkesh, Aidan, Sam]
We did a tap-tap-tappy-tap test of the OMC to try to find its resonances. We looked at some combination of the PDH error signal and the DCPD signal in a couple of different noise configurations. The data included below shows tapping of the major tombstone objects as well the breadboard. I don't see any strong evidence of resonances below the very sharp resonance at 1300 Hz (which I interpret as the diving board mode of the breadboard). If I get free, I 'll post some plots of the different breadboard resonances you can excite by tapping in different places.
(The "normalized" tapping response is abs(tap - reference)./reference.) |
87
|
Fri Nov 9 00:23:12 2007 |
pkp | Update | OMC | X and Z resonances | I got a couple of resonance plots going for now. I am still having trouble getting the Y measurement going for some reason. I will investigate that tommorow. But for tonight and tommorow morning, here is some food for thought. I have attached the X and Z transfer functions below. I compared them to Norna's plots - so just writing out what I was thinking -
Keep in mind that these arent high res scans and have been inconviniently stopped at 0.5 Hz .
Z case --
I see two small resonances and two large ones - the large ones are at 5.5 Hz and 0.55 Hz and the small ones at 9 Hz and 2 Hz respectively. In Norna's resonances, these features arent present. Secondly, the two large peaks in Norna's measurement are at 4.5 Hz and just above 1 Hz. Which was kind of expected, since we shortened the wires a bit, so one of the resonances moved up and I suppose that the other one moved down for the same reason.
X case --
Only one broad peak at about 3 Hz is seen here, whereas in Norna's measurement, there were two large peaks and one dip at 0.75 Hz and 2.5 Hz. I suspect that the lower peak has shifted lower than what I scanned to here and a high res scan going upto 0.2 Hz is taking place overnight. So we will have to wait and watch.
Pitch Roll and Yaw can wait for the morning. |
93
|
Mon Nov 12 10:53:58 2007 |
pkp | Update | OMC | Vertical Transfer functions | [Norna Sam Pinkesh]
These plots were created by injected white noise into the OSEMs and reading out the response of the shadow sensors ( taking the power spectrum). We suspect that some of the additional structure is due to the wires. |
99
|
Wed Nov 14 07:48:38 2007 |
norna | Omnistructure | OMC | OMC Cable dressing | [Snipped from an email]
1) Last Friday Pinkesh and I set the OMC up with only the top three OSEMs and took a vertical transfer function. We had removed the other OSEMs due to difficulty of aligning all OSEMs with the weight of the bench etc bringing the top mass lower than the tablecloth can accommodate. See attached TF.Clearly there are extra peaks (we only expect two with a zero in between) and my belief is that at least some of them are coupling of other degrees of freedom caused by the electrical wiring. Pinkesh and I also noticed the difficulty of maintaining alignment if cables got touched and moved around. So.....
2) Yesterday Dennis and I took a look at how much moving a cable bundle around (with the peak shielding) changed the DC alignment. In a not too precise experiment ( using HeNe laser reflecting off the bench onto a surface ~ 1 metre away) we saw that we could reposition the beam one or two mm in yaw and pitch. This corresponds to ~ one or two mrad which is ~ the range of the OSEM DC alignment. We discussed possibility of removing the cabling from the middle mass, removing the peak and taking it from the bench directly to the structure above. I asked Chub if he could make an equivalent bundle of wires as those from the two preamps to see what happens if we repeat the "moving bundle" experiment. So...
3) Today Chub removed the cabling going to the preamps and we replaced it with a mock up of wire bundle going directly from the preamps to the structure above. See attached picture. The wires are only attached to the preamp boxes weighted down with masses but the bundle is clamped at the top. We repeated the "wiggle the bundle" test and couldnt see any apparent movement ( so maybe it is at most sub-mm). The cable bundle feels softer.
The next thing Chub did was to remove the second bundle ( from photodiodes, heater, pzt) from its attachment to the middle mass and strip off the peek. It is now also going to the top of the structure directly. The whole suspension now appears freer. We discussed with Dennis the "dressing " of the wires. There are some minor difficulties about how to take wires from the bright side to the dark side, but in general it looks like that the wires forming the second "bundle" could be brought to the "terminal block" mounted on the dark side and from there looped up to the top of the structure. We would have to try all this of course to see the wiring doesnt get in the way of other things (e.g. the L and R OSEMs). However this might be the way forward. So...
4) Tomorrow Pinkesh and I will check the alignment and then repeat the vertical transfer function measurement with the two bundles as they are going from bench to top of structure. We might even do a horizontal one if the middle mass is now within range of the tablecloth.
We can then remove preamp cables completely and lay the second bundle of cables on the optical bench and repeat the TFs.
The next thing will be to weigh the bench plus cables. This will allow us to
a) work out what counterbalance weights are needed - and then get them manufactured
b) firm up on how to handle the extra mass in terms of getting the masses at the correct height.
And in parallel Chub will work on the revised layout of cabling.
Looking a little further ahead we can also get some stiffness measurements made on the revised bundle design ( using Bob's method which Alejandro also used) and fold into Dennis's model to get some sanity check the isolation.
I think that's it for now. Comments etc are of course welcome.
Norna |
102
|
Wed Nov 14 16:54:54 2007 |
pkp | Update | OMC | Much better looking vertical transfer functions | [Norna Pinkesh]
So after Chub did his wonderful mini-surgery and removed the peek from the cables and after Norna and I aligned the whole apparatus, the following are the peaks that we see.
It almost exactly matches Norna's simulations and some of the extra peaks are possibly due to us exciting the Roll/longitudnal/yaw and pitch motions. The roll resonance is esp prominent.
We also took another plot with one of the wires removed and will wait on Chub before we remove another wire. |
105
|
Thu Nov 15 17:09:37 2007 |
pkp | Update | OMC | Vertical Transfer functions with no cables attached. | [Norna Pinkesh]