40m QIL Cryo_Lab CTN SUS_Lab CAML OMC_Lab CRIME_Lab FEA ENG_Labs OptContFac Mariner WBEEShop
  40m elog, Page 242 of 357  Not logged in ELOG logo
ID Date Author Type Categoryup Subject
  1495   Sun Apr 19 03:34:05 2009 YoichiUpdateLockingSaturday night lock
Tonight I was able to go up to arm power = 33, by mainly tweaking the DARM gain. A small progress.
In order to give more phase margin to the CARM MC_L path, I added a 300:100 filter to C1:LSC-MC.
To reduce the load to the lsc computer I deleted several filters from the filter bank, which were not used in the locking scripts.
Before I deleted the filters, I checked in the current chans directory into the svn repository.
If you want to restore the deleted filters, go back to the revision 36142.
  1498   Mon Apr 20 05:18:42 2009 YoichiConfigurationLockingFM6 and FM10 of LSC-MC were restored
During tonight's locking work, I realized that FM6 and FM10 (both resonant gains around 20Hz) were actually activated by cm_step.
So I restored those filters from the svn history.
Instead, I removed a bunch of unused filters from LSC-DEMOD and LSC-DEMOD_A (moving zero filters) to off load c1lsc.

As for the locking itself, the DARM loop becomes unstable at around arm power = 30. I may have to add a filter to give a broader phase bubble.
  1500   Mon Apr 20 18:17:44 2009 robSummaryLockingCARM offset/Power rubric

Plotted assuming the average arm power goes up to ~80.  No DARM offset.

  1509   Thu Apr 23 16:27:24 2009 YoichiUpdateLockingLocking with the cryo-pump
The last night, the IFO was unstabler than usual and the locking script often failed before reaching the power up stage.
The failure happened at random points.
I'm not sure if this is related to the operation of the cryo-pump.
The mode cleaner reflection image seemed to move around more than usual. Maybe it was just a high seismic night.
  1514   Fri Apr 24 03:57:30 2009 YoichiUpdateLockingDARM demod phase
Tonight, I was able to go up to arm power = 40 by tweaking the DARM demodulation phase.
I think the DARM loop became unstable because the demodulation phase was not right and the error signal contained some junk from I-phase.
I did not do any sophisticated demodulation phase optimization. Rather I just tweaked the phase so that the dark port image becomes stable.
I will do more careful demodulation phase tuning next time.
  1515   Fri Apr 24 04:38:49 2009 YoichiUpdateLockingDARM demod phase

Quote:
Tonight, I was able to go up to arm power = 40 by tweaking the DARM demodulation phase.
I think the DARM loop became unstable because the demodulation phase was not right and the error signal contained some junk from I-phase.
I did not do any sophisticated demodulation phase optimization. Rather I just tweaked the phase so that the dark port image becomes stable.
I will do more careful demodulation phase tuning next time.


In the next try, I was actually able to go up to arm power = 70 stably.
At this power level we are ready for the RF CARM hand off.
  1516   Fri Apr 24 11:34:32 2009 robUpdateLockingDARM demod phase

Quote:

Quote:
Tonight, I was able to go up to arm power = 40 by tweaking the DARM demodulation phase.
I think the DARM loop became unstable because the demodulation phase was not right and the error signal contained some junk from I-phase.
I did not do any sophisticated demodulation phase optimization. Rather I just tweaked the phase so that the dark port image becomes stable.
I will do more careful demodulation phase tuning next time.


In the next try, I was actually able to go up to arm power = 70 stably.
At this power level we are ready for the RF CARM hand off.


There's actually code in place in the LSC to dynamically adjust the demod phase for AS1. I've never made much use of it, because it's possible to get around the problem with some gain tweaking if you start at the right phase, or because I did the DC readout handoff earlier.

Attached is a cartoon showing how the demod phase at the dark port changes as the CARM offset is decreased.
  1519   Fri Apr 24 17:26:57 2009 YoichiUpdateLockingDARM demod phase

Quote:

There's actually code in place in the LSC to dynamically adjust the demod phase for AS1. I've never made much use of it, because it's possible to get around the problem with some gain tweaking if you start at the right phase, or because I did the DC readout handoff earlier.

Attached is a cartoon showing how the demod phase at the dark port changes as the CARM offset is decreased.


The cartoon is very nice.
I actually changed the demod phase continuously as the CARM offset was reduced to get up to arm power = 70.
As the CARM offset is changed, not only the DARM signal gain but also the phase margin around 100Hz changes if you use a fixed demodulation phase.
So it was necessary to change the demodulation phase to keep the DARM loop stable.
  1522   Sat Apr 25 03:27:34 2009 YoichiUpdateLockingLocking status
Yoichi, Peter,

We are working on the final step of the lock acquisition, RF CARM hand off.
I was able to hand off the CARM error signal to RF once, but lost lock when decreasing the CARM offset to zero (it was too rapid).
I will try to make the process more robust tomorrow.
  1523   Sun Apr 26 02:13:18 2009 YoichiUpdateLockingTwo more successes of RF CARM handoff
Tonight, the RF CARM hand off (mostly) succeeded twice.
But still, the IFO lost lock when I reduced the REFL_DC gain in the AO path to zero.

At the beginning of tonight's work, MICH DD hand off failed several times. This was because the the PD9 gains were set to zero.
I found that the offset script, which I called before starting the locking, fails to restore the gain values sometimes.
This happens when ezcaread fails to read the current gain. We have to be careful when running the LSCoffsets script.
  1526   Tue Apr 28 04:30:16 2009 YoichiUpdateLockingRF full lock
Yoichi, Peter

I believe we have succeeded in the full lock of the interferometer with the RF signals.
The lock process is reasonably robust and repeatable.

I did a scan of the RF CARM offset and plotted the arm power as a function of the CARM offset (see the attachment).
The arm power goes maximum at non-zero CARM offset. I guess the RF CARM error signal has some offset.
Maybe the demodulation phase is wrong ? I will tweak this tomorrow.
The script to do this scan can be found at /cvs/cds/caltech/scripts/CM/CARMSweep.

I haven't tried DC readout yet.
  1529   Tue Apr 28 16:36:24 2009 robHowToLockingsetting the RF CARM demod phase

To set the demod phase for RF CARM, sensed at REFL2 (REFL 166I), it suffices to set the demod phase for REFL2 to be the optimal phase for controlling SRCL in a no-arm state.

  1530   Tue Apr 28 17:51:13 2009 robHowToLockingsetting the RF CARM demod phase

Quote:

To set the demod phase for RF CARM, sensed at REFL2 (REFL 166I), it suffices to set the demod phase for REFL2 to be the optimal phase for controlling SRCL in a no-arm state.

 

For POX33, the ideal phase for single arm locking does not yield a zero-offset CARM signal.  So the offset needs to be manipulated digitally. 

  1531   Wed Apr 29 04:03:51 2009 YoichiUpdateLockingCARM RF changed to REFL_2I
Yoichi, Peter

As Rob suggested, the optimal demodulation phase is easier to find for REFL_2I than POX_1I.
Moreover, for 166MHz LO, we have a phase shifter (delay line) already installed. So we can easily change the demodulation phase of REFL_2I.
Tonight, we switched the RF CARM signal to REFL_2I.
To do so, I changed the signal going to the REFL1 input of the common mode board from POX_1I to REFL_2I.
I moved a BNC-T installed at the output of POX_1I to the REFL_2I output to split the REFL_2I signal and send it to the CM board.
Since the gain of the REFL_2I was about 20dB lower than that of POX_1I, I increased the gain of the SR560, which is installed between the REFL_2 demodulation board and the CM board, from 1 to 10.

With some gain tweaks, we were able to hand off the CARM from REFL_DC to REFL_2I. We also succeeded in switching the REFL_2I ADC channel from PD11 to PD2_DC (the output of the length path from the CM board). This switching is necessary in order to engage the boost on the CM board.

There remains some offset in the CARM when the arm power is maximized. This is expected because the REFL_2I demodulation phase is probably not exactly right.
I will optimize the demodulation phase tomorrow.
  1533   Wed Apr 29 15:56:43 2009 robUpdateLockingeffect of SRCL detune on ARM powers in a CARM sweep

With no DARM offset, sweeping CARM shows an asymmetry between the state where we lock to a DARM spring and the state with a DARM anti-spring.  This is why we have a link between the DARM and CARM optical springs. 

For each DARM detune direction (positive or negative, spring or anti-spring), there is only one CARM direction which can yield a DC-based error signal lock with a CARM offset but no DARM offset, which is what we want.

  1534   Thu Apr 30 05:49:06 2009 YoichiUpdateLocking166MHz LO phase changed
In order to optimize the REFL_2I demod phase, I changed the delay line setting for the 166MHz LO.
Right now, the delay is not yet optimal.
Since the AS166 shares the same LO, the digital demodulation phase of the AS166 had to be changed too.
The DD demod phases and the DD hand off script were also tweaked to improve the resonant condition of the central part.
Now we have more 166MHz coming out of the AS port and the SPOB is larger (more 33MHz resonant in PRC).

Since REFL166 and AS166 demodulation phases are not yet optimized, the cm_step script won't work at this moment.
  1535   Thu Apr 30 15:10:54 2009 robUpdateLockingCARM RF changed to REFL_2I

Quote:
Yoichi, Peter

As Rob suggested, the optimal demodulation phase is easier to find for REFL_2I than POX_1I.
Moreover, for 166MHz LO, we have a phase shifter (delay line) already installed. So we can easily change the demodulation phase of REFL_2I.
Tonight, we switched the RF CARM signal to REFL_2I.
To do so, I changed the signal going to the REFL1 input of the common mode board from POX_1I to REFL_2I.
I moved a BNC-T installed at the output of POX_1I to the REFL_2I output to split the REFL_2I signal and send it to the CM board.
Since the gain of the REFL_2I was about 20dB lower than that of POX_1I, I increased the gain of the SR560, which is installed between the REFL_2 demodulation board and the CM board, from 1 to 10.

With some gain tweaks, we were able to hand off the CARM from REFL_DC to REFL_2I. We also succeeded in switching the REFL_2I ADC channel from PD11 to PD2_DC (the output of the length path from the CM board). This switching is necessary in order to engage the boost on the CM board.

There remains some offset in the CARM when the arm power is maximized. This is expected because the REFL_2I demodulation phase is probably not exactly right.
I will optimize the demodulation phase tomorrow.



From Optickle simulations, it looks like the SRCL/CARM gain ratio at REFL I2 is about 8e-4. So a 1 nanometer offset in SRCL yields 0.8 picometers of offset in CARM.
  1536   Fri May 1 01:32:43 2009 YoichiUpdateLocking166MHz LO phase adjustment
I continued to adjust the REFL_2I demodulation phase.
I first optimized the demod phase for SRCL in the DRMI configuration (the error signals were DDs).
Then I restored the full IFO and offset locked it.
Before handing the DARM to RF, I adjusted the 166MHz delay line to maximize the SRCL signal at REFL_2I.
I did this before the DARM RF hand off because changing the delay line setting also changes the AS166 demodulation phase.
After this, I adjusted the digital phase shifter for AS166 to maximize the DARM signal for this port.

I also adjusted the digital demodulation phase of PD11 (REFL_2I) because the optimal demodulation phase for the initial lock acquisition is somewhat (15deg)
different from the optimal demodulation phase for the SRCL when the central part is locked with the DD signals.
This happens because the resonant condition of the central part (lock points of the recycling cavities) changes when the error signals are switched to the DD signals,
due to the offset in the DD signals. This is not good and should be fixed by the optimization of the DD demodulation phases.

Finally, I reduced the CARM offset to zero and tweaked the delay line a bit to maximize the arm power.

Right now, the locking script runs fine until the end.
At the end of the script, I was able to engage the boost on the CM board.
  1537   Fri May 1 10:04:10 2009 robUpdateLocking166MHz LO phase adjustment

Quote:
I continued to adjust the REFL_2I demodulation phase.
I first optimized the demod phase for SRCL in the DRMI configuration (the error signals were DDs).
Then I restored the full IFO and offset locked it.
Before handing the DARM to RF, I adjusted the 166MHz delay line to maximize the SRCL signal at REFL_2I.
I did this before the DARM RF hand off because changing the delay line setting also changes the AS166 demodulation phase.
After this, I adjusted the digital phase shifter for AS166 to maximize the DARM signal for this port.

I also adjusted the digital demodulation phase of PD11 (REFL_2I) because the optimal demodulation phase for the initial lock acquisition is somewhat (15deg)
different from the optimal demodulation phase for the SRCL when the central part is locked with the DD signals.
This happens because the resonant condition of the central part (lock points of the recycling cavities) changes when the error signals are switched to the DD signals,
due to the offset in the DD signals. This is not good and should be fixed by the optimization of the DD demodulation phases.

Finally, I reduced the CARM offset to zero and tweaked the delay line a bit to maximize the arm power.

Right now, the locking script runs fine until the end.
At the end of the script, I was able to engage the boost on the CM board.



Awesome. Up next: dewhitening.
  1541   Sun May 3 22:48:12 2009 YoichiUpdateLockingSome measurements at the lock point
I attached some measurement results at when the IFO is at the full lock point.

The first plot shows the trend of the arm powers after the interferometer was locked.
The arm powers slowly increased after the lock. This increase is observed every time the IFO is locked.
Probably this is some sort of a thermal effect (mirror lensing, PD efficiency etc).

The second plot is a CARM offset sweep. Even after the demodulation phase optimization, the lock point is not exactly at the resonance.

The third plot is the open loop TF of the AO path. The CM loop UGF is about 20kHz.
The boost and the superboost1 were turned on when this TF was measured. The IFO loses lock if the superboost2 is turned on.

TO DO LIST
Measured the DARM loop shape.
I could not turn on the dewhitening filter for ETMY. ETMX had no trouble. I will check the dewhitening circuit.
  1544   Tue May 5 05:16:12 2009 YoichiUpdateLockingDC Readout and DARM response
Tonight, I was able to switch the DARM to DC readout a couple of times.
But the lock was not as stable as the RF DARM. It lost lock when I tried to measure the DARM loop gain.

I also measured DARM response when DARM is on RF.
The attached plot shows the DARM optical gain (from the mirror displacement to the PD output).
The magnitude is in an arbitrary unit.

I measured a transfer function from DARM excitation to the DARM error signal. Then I corrected it for the DARM open loop gain and the pendulum response to get the plot below.

There is an RSE peak at 4kHz as expected. The origin of the small bump and dip around 2.5kHz and 1.5kHz are unknown.
I will consult with the Optickle model.
I don't know why the optical gain decreases below 50Hz (I don't think it actually decreases).
Seems like the DARM loop gain measured at those frequencies are too low.
I will retry the measurement.
  1545   Tue May 5 08:26:56 2009 robUpdateLockingDC Readout and DARM response

Quote:
Tonight, I was able to switch the DARM to DC readout a couple of times.
But the lock was not as stable as the RF DARM. It lost lock when I tried to measure the DARM loop gain.

I also measured DARM response when DARM is on RF.
The attached plot shows the DARM optical gain (from the mirror displacement to the PD output).
The magnitude is in an arbitrary unit.

I measured a transfer function from DARM excitation to the DARM error signal. Then I corrected it for the DARM open loop gain and the pendulum response to get the plot below.

There is an RSE peak at 4kHz as expected. The origin of the small bump and dip around 2.5kHz and 1.5kHz are unknown.
I will consult with the Optickle model.
I don't know why the optical gain decreases below 50Hz (I don't think it actually decreases).
Seems like the DARM loop gain measured at those frequencies are too low.
I will retry the measurement.


The optical gain does decrease below ~50Hz--that's the optical spring in action. The squiggles are funny. Last time we did this we measured the single arm TFs to compensate for any tough-to-model squiggles in the transfer functions which might arise from electronics or the suspensions.
  1548   Tue May 5 11:44:33 2009 robUpdateLocking DARM response

Here's the RF DARM optical response, on the anti-spring side, from optickle. Note that for the f1 sideband, changing the demod phase mostly adjusts the overall gain, while for the f2 sideband a change in demod phase alters the shape of the response. This is the quadrature-selecting power of using a single RF sideband as a local oscillator.
  1550   Wed May 6 02:39:20 2009 YoichiHowToLockingHow to go to DC readout
I wrote a script called DC_readout, which you can find in /cvs/cds/caltech/scripts/DRFPMI/bang/nospring/.

Currently, the locking script succeeds 1/3 of the time. The freaky parts are the MC_F hand off and REFL_DC hand off.
MC_F hand off succeeds 70% of the time. REFL_DC goes well about a half of the time. Combined, the success rate is about 1/3.
We need some work on those hand offs.
Once you pass those freaky parts, the cm_step script usually goes smoothly and you will reach the full RF lock with the boost and the super boost1 engaged on the CM board.

To go to DC readout from there, run the DC_readout script.
First, this script will put some offset to the DARM loop so that some carrier light will leak to the AS port.
You are prompted to lock the OMC. Move the OMC length offset slider to find the carrier resonance and lock the OMC.
You have to make sure that it is carrier, not the 166MHz sideband. Usually the carrier light pulsates around 10Hz or so whereas the 166MHz SB is stable.
Once you locked the OMC to the carrier, hit enter on the terminal running the DC_readout script.
The script will do the rest of the hand off.
Once the script has finished, you may want to check darm_offset_dc in the C1LSC_LA_SET screen. This value sets the DC offset (a.k.a. the homodyne phase).
You may want to change it to what you want.
  1557   Thu May 7 18:12:12 2009 peteUpdateLockingarm power curve

I've plotted TRX, TRY, PD12I and PD11Q.  Arm powers after locking increase for a few tens of minutes, peak out, and then decrease before lock is lost.

 

 

  1558   Thu May 7 23:21:04 2009 peteUpdateLockingarm power curve

Quote:

I've plotted TRX, TRY, PD12I and PD11Q.  Arm powers after locking increase for a few tens of minutes, peak out, and then decrease before lock is lost.

 

 

 I should have mentioned that the AS port camera image seems to get progressively uglier over the course of these locks.  Maybe we can use the JoeCam to make a movie of it. 

  1560   Fri May 8 02:08:59 2009 peteUpdateLockinglock stretches

locks last for about an hour.  this was true last night as well (see "arm power curve" entries).   the second lock shown here evolves differently for unknown reasons.  the jumps in the arm powers of the first lock are due to turning on DC readout.  length-to-angle needs tuning.

 

 

  1561   Fri May 8 02:39:02 2009 pete, ranaUpdateLockingcrossover

attached plot shows MC_IN1/MC_IN2.  needs work.

This is supposed to be a measurement of the relative gain of the MCL and AO paths in the CM servo. We expect there to

be a more steep slope (ideally 1/f). Somehow the magnitude is very shallow and so the crossover is not stable. Possible

causes? Saturations in the measurement, broken whitening filters, extremely bad delay in the digital system? needs work.

 

  1565   Fri May 8 15:40:44 2009 peteUpdateLockingprogressively weaker locks

the align script was run after the third lock here.  it would have been interesting to see the arm powers in a 4th lock 

  1585   Thu May 14 02:36:05 2009 peteUpdateLockingunstable IFO

It seems that the MC3 problem is intermittent (one-day trend attached).  I tried to take advantage of a "clean MC3" night, but the watch script would usually fail at the transition to DC CARM and DARM.  It got past this twice and then failed later, during powering up.   I need to check the handoff.

 

  1611   Wed May 20 01:53:48 2009 rob, peteUpdateLockingviolin mode filters in drstep_bang

Recently the watch script was having difficulty grabbing a lock for more than a few seconds.  Rob discovered that the violin notch filters which were activated in the script were causing the instability.  We're not sure why yet.  The script seems significantly more stable with that step commented out.

  1641   Tue Jun 2 02:28:58 2009 peteUpdateLockingDD handoff work

alberto, pete

 

We worked on tuning the DD handoff tonight.  We checked the DD PD alignments and they looked fine.  First I tuned the 3 demod phases to minimize offsets.  Then I noticed that the post-handoff MICH xfer function needed an increase in gain to look like the pre-handoff xfer function (which has a UGF of about 25 Hz).  I increased the MICH PD9_Q gain from 2 to 7 in the input matrix.   But, the handoff to PRC still failed, so tomorrow we will try to find out why.

In the plot, ref0 is before MICH handoff, and ref1 is after MICH handoff.  There is also a PRC trace (before PRC handooff).

 

 

  1645   Wed Jun 3 03:22:16 2009 peteUpdateLockingDD handoff

Rana, Alberto, Pete

We have the DD handoff nominally working.  Sometimes, increasing the SRC gain at the end makes MICH get unstable.  This could be due to a non-diagonal term in the matrix, or possibly because the DRM locks in a funky mode sometimes. 

To get the DD handoff working, first we tuned demod phases in order to zero the offsets in the PD signals handed-off-to.  Based on transer function measurements, I set the PRC PD6_I element to 0.1, and set the PD8_I signal to 0, since it didn't seem to be contributing much.  We also commented out the MICH gain increase at the end of the DD_handoff script.

It could still be more stable, but it seems to work most of the time.

 

 

  1652   Thu Jun 4 16:54:19 2009 peteUpdateLockingdaytime DD handoff

I played with the DD handoff during the day.  The DRM dark port was flickering like a candle flame in Dracula's castle.  The demod offsets for the handoff signals looked fine.  After MICH handoff, the MICH_CTRL started to get unstable at some low frequency, maybe 3 Hz (I didn't measure).  So I increased the MICH gain from 0.1 to 0.17 and it settled down.  PRC and SRC went fine.  Then the DD_handoff script raised the MICH gain to 0.7, and an instability started to grow in MICH_CTRL (at some higher frequency).  I decreased the MICH gain from 0.7 to 0.5, and it settled down and stayed stable.

  1654   Fri Jun 5 01:10:13 2009 rob, peteUpdateLockingundermined

We were stymied early in the evening by a surreptitiously placed, verbo-visually obfuscated command in the drstep script. 

  1655   Fri Jun 5 02:59:03 2009 pete, albertoUpdateLockingtdsavg failure in cm_step script

the command

tdsavg 5 C1:LSC-PD4_DC_IN1

was causing grievous woe in the cm_step script.  It turned out to fail intermittently at the command line, as did other LSC channels.  (But non-LSC channels seem to be OK.)  So we power cycled c1lsc (we couldn't ssh).

Then we noticed that computers were out of sync again (several timing fields said 16383 in the C0DAQ_RFMNETWORK screen).  We restarted c1iscey, c1iscex, c1lsc, c1susvme1, and c1susvme2.  The timing fields went back to 0.  But the tdsavg command still  intermittently said "ERROR: LDAQ - SendRequest - bad NDS status: 13".

The channel C1:LSC-SRM_OUT16 seems to work with tdsavg every time.

Let us know if you know how to fix this. 

 

  1656   Fri Jun 5 13:51:49 2009 robUpdateLockingtdsavg failure in cm_step script

Quote:

the command

tdsavg 5 C1:LSC-PD4_DC_IN1

was causing grievous woe in the cm_step script.  It turned out to fail intermittently at the command line, as did other LSC channels.  (But non-LSC channels seem to be OK.)  So we power cycled c1lsc (we couldn't ssh).

Then we noticed that computers were out of sync again (several timing fields said 16383 in the C0DAQ_RFMNETWORK screen).  We restarted c1iscey, c1iscex, c1lsc, c1susvme1, and c1susvme2.  The timing fields went back to 0.  But the tdsavg command still  intermittently said "ERROR: LDAQ - SendRequest - bad NDS status: 13".

The channel C1:LSC-SRM_OUT16 seems to work with tdsavg every time.

Let us know if you know how to fix this. 

 

 Did you try restarting the framebuilder?

 

What you type is in bold:

op440m> telnet fb40m 8087

daqd> shutdown

 

  1658   Fri Jun 5 17:22:55 2009 peteUpdateLockingdaytime locking

After fixing the tp problem, I tried locking again.  Grabbing and DD handoff, no problem.  Died earlier than last night, handing off CARM to REFL_DC, around arm power of 4 or so.  Seems to happen after turning off the moving zero, Rob says it might be touchy in daytime.

  1659   Sat Jun 6 01:44:53 2009 rob UpdateLocking?

Lock acquisition is proceeding smoothly for the most part, but there is a very consistent failure point near the end of the cm_step script.

Near the end of the procedure, while in RF common mode, the sensing for the MCL path of the common mode servo is transitioned from a REFL 166I signal which comes into the LSC whitening board from the demodulator, to another copy of the signal which has passed through the common mode board, and is coming out of the Length output of the common mode board.  We do this because the signal which comes through the CM board sees the switchable low-frequency boost filter, and so both paths of the CM servo (AO and MCL) can get that filter switched on at the same time.

The problem is occurring after this transition, which works reliably.  However, when the script tries to remove the final CARM offset, and bring the offset to zero, lock is abruptly lost.  DARM, CM, and the crossover all look stable, and no excess noise appears while looking at the DARM, CARM, MCF spectra.  But lock is always lost right about the same offset. 

Saturation somewhere?

  1660   Sun Jun 7 04:57:39 2009 YoichiUpdateLocking?

Quote:

Lock acquisition is proceeding smoothly for the most part, but there is a very consistent failure point near the end of the cm_step script.

Near the end of the procedure, while in RF common mode, the sensing for the MCL path of the common mode servo is transitioned from a REFL 166I signal which comes into the LSC whitening board from the demodulator, to another copy of the signal which has passed through the common mode board, and is coming out of the Length output of the common mode board.  We do this because the signal which comes through the CM board sees the switchable low-frequency boost filter, and so both paths of the CM servo (AO and MCL) can get that filter switched on at the same time.

The problem is occurring after this transition, which works reliably.  However, when the script tries to remove the final CARM offset, and bring the offset to zero, lock is abruptly lost.  DARM, CM, and the crossover all look stable, and no excess noise appears while looking at the DARM, CARM, MCF spectra.  But lock is always lost right about the same offset. 

Saturation somewhere?

 I've seen this before. At that time, the problem was gone spontaneously the next day.

You could stop just before the offset reaches zero and then try to slowly reduce the offset manually to see where is the threshold.

 

  1663   Tue Jun 9 23:25:24 2009 robUpdateLocking?

Quote:

Quote:

Lock acquisition is proceeding smoothly for the most part, but there is a very consistent failure point near the end of the cm_step script.

Near the end of the procedure, while in RF common mode, the sensing for the MCL path of the common mode servo is transitioned from a REFL 166I signal which comes into the LSC whitening board from the demodulator, to another copy of the signal which has passed through the common mode board, and is coming out of the Length output of the common mode board.  We do this because the signal which comes through the CM board sees the switchable low-frequency boost filter, and so both paths of the CM servo (AO and MCL) can get that filter switched on at the same time.

The problem is occurring after this transition, which works reliably.  However, when the script tries to remove the final CARM offset, and bring the offset to zero, lock is abruptly lost.  DARM, CM, and the crossover all look stable, and no excess noise appears while looking at the DARM, CARM, MCF spectra.  But lock is always lost right about the same offset. 

Saturation somewhere?

 I've seen this before. At that time, the problem was gone spontaneously the next day.

You could stop just before the offset reaches zero and then try to slowly reduce the offset manually to see where is the threshold.

 

 

Well, it hasn't gone away yet.  It happened Sat, Mon, and Tues afternoon, as well as Friday.  The threshold varies slightly, but is always around ~200-300 cnts.   I've tried reducing the offset with the signal coming from the CM board and the signal not going through the CM board, I've also tried jumping the signal to zero (rather than a gradual reduction). 

Tonight we'll measure the MC length and set the modulation frequencies, and maybe try some MZ tweaking to do RFAMMon minimization.    

  1675   Tue Jun 16 02:09:31 2009 robUpdateLockingDD handoff finally working

I had trouble getting the SRC handoff from SD to DD to work.  I tried different gains, flipping the PD7 & 8 demod phases by 180 degrees, and messing with the output matrix to reduce cross-couplings in the state with MICH & PRC on DD and SRC on SD.  Eventually I decided to try to make the DRM matrix diagonalization work. 

It does, mostly.  The handoff is now stable, and the loops all have UGFs around 100Hz.  So, tonight anyways, it's possible to run senseDRM and then loadDRMImatrixData.m and run the resulting tdswrite command, and have a working handoff.  I had to eliminate a few PDs (PD5 & PD10) to get it to work properly. 

It would be nice if this script would measure all the PDs and allow individual setting of loop UGFs and measurement frequencies. 

 

 

  1676   Tue Jun 16 03:43:04 2009 robUpdateLockingsame troubles

Lock is still being lost, right at the end of the process when trying to reduce the CARM offset to zero.

  1678   Tue Jun 16 14:02:16 2009 robUpdateLockinglocked

The IFO is locked, at the operating point (zero CARM offset).  The problem with reducing the residual CARM offset in the last stage appears to have been because the common mode gain was getting too high, and so the loop was going unstable at high frequencies. 

The cm_step script is currently a confusing mess, with all the debugging statements.  I'll clean it up this afternoon and check that it still works.

  1679   Tue Jun 16 16:10:01 2009 peteUpdateLockinginput matrix experiments

Last night Rob ran senseDRM and loadDRMImatrixData and came up with the following for the input matrix:

tdswrite C1:LSC-ITMTRX_b2 0.065778 \
C1:LSC-ITMTRX_d2 2.2709 \
C1:LSC-ITMTRX_f2 2.9361 \
C1:LSC-ITMTRX_122 0.42826 \
C1:LSC-ITMTRX_b3 -0.064839 \
C1:LSC-ITMTRX_d3 -0.016913 \
C1:LSC-ITMTRX_f3 -0.021576 \
C1:LSC-ITMTRX_123 -0.0025243 \
C1:LSC-ITMTRX_b5 0.3719 \
C1:LSC-ITMTRX_d5 1.3109 \
C1:LSC-ITMTRX_f5 -0.16412 \
C1:LSC-ITMTRX_125 0.39574 \
C1:LSC-ITMTRX_33 0 \
C1:LSC-ITMTRX_42 0 \
C1:LSC-ITMTRX_155 0

Today, I reran these and got the following, and DD_handoff remained happy:

tdswrite C1:LSC-ITMTRX_b2 -0.10329 \
C1:LSC-ITMTRX_d2 2.0344 \
C1:LSC-ITMTRX_f2 3.2804 \
C1:LSC-ITMTRX_122 0.22516 \
C1:LSC-ITMTRX_b3 -0.076292 \
C1:LSC-ITMTRX_d3 -0.014603 \
C1:LSC-ITMTRX_f3 -0.12101 \
C1:LSC-ITMTRX_123 0.0054128 \
C1:LSC-ITMTRX_b5 0.33521 \
C1:LSC-ITMTRX_d5 1.1425 \
C1:LSC-ITMTRX_f5 -0.32759 \
C1:LSC-ITMTRX_125 0.25877 \
C1:LSC-ITMTRX_33 0 \
C1:LSC-ITMTRX_42 0 \
C1:LSC-ITMTRX_155 0

I wanted to remeasure with the canonical output matrix (-0.7 from MICH to PRM and 0.7 from MICH to SRM), but the DRM freaked out when MICH to PRM went below -0.3.

  1680   Tue Jun 16 17:38:35 2009 robUpdateLockingDeWhites ON

With the common mode servo bandwidth above 30kHz and the BOOST on (1), I was able to switch on the test mass dewhitening.  Finally.

  1684   Thu Jun 18 23:08:46 2009 robUpdateLockingspectrum

Here's a noise spectrum of the RSE interferometer, in anti-spring mode, with RF readout.  I'd say the calibration is "loose."

I used the Buonanno & Chen modification of the KLMTV IFO transfer functions to model the DARM opto-mechanical response.  I just guessed at the quadrature, and normalized the optical gain at the frequency of the calibration line used (927Hz, not visible on the plot).

  1709   Tue Jun 30 23:09:40 2009 AlbertoUpdateLockingchronicles of some locking attempts

Tonight I tried to lock the interferometer. At the first attempts the arm power didn't go above about 4. The mode cleaner seemed to be not well aligned and it lost lock or got stuck on a wrong mode. I had to run the MC_UP and MC_DOWN scripts to lock it again.

After that the locking proceed more smoothly; at least till a power level in the arms of about 60. Then again the mode cleaner lost lock and I had to run the scripts again. Without the MCWFS servo off the MC reflected power is still rather high (about 1.7); also even when the WFS servo is engaged the reflected power is about 0.5, versus 0.3 that it should be.

Those are both signs of a not very good alignment. Tomorrow I'll have to work on the injection periscope on the PSL table to try to fix that.

  1742   Tue Jul 14 00:57:11 2009 AlbertoUpdateLockingphotodiode alignment check

Since lately the alignment of the input beam to the interferometer has changed, I went checking the alignment of the beam on the photodiodea. They were all fine except for pd9, that is AS DD 199. Here the DC is totally null. The beam seems to go right on the diode but the scope on the PD's DC output shows no power. This is really strange and bad.

  1747   Wed Jul 15 11:38:31 2009 robUpdateLockingMC_F channel dead

It's railed.  This is what halted locking progess on Monday night, as this channel is used for the offloadMCF script, which slowly feeds back a CARM signal to the ETMs to prevent the VCO from saturating.

 

Attached is a 5 day trend, which shows that the channel went dead a few days ago.  All the channels shown are being collected from the same ICS110B (I think), but only some are dead.  It looks like they went dead around the time of the "All computers down" from Sunday.

ELOG V3.1.3-