  12657   Fri Dec 2 11:56:42 2016 gautamUpdateLSCMC1 LEMO jiggled

I noticed 2 periods of frequent IMC locklosses on the StripTool trace, and so checked the MC1 PD readout channels to see if there were any coincident glitches. Turns out there wasnt yes BUT - the LR and UR signals had changed significantly over the last couple of days, which is when I've been working at 1X5. The fast LR readback was actually showing ~0, but the slow monitor channel had been steady so I suspected some cabling shenanigans.

Turns out, the problem was that the LEMO connector on the front of the MC1 whitening board had gotten jiggled ever so slightly - I re-jiggled it till the LR fast channel registered similar number of counts to the other channels. All looks good for now. For good measure, I checked the 3 day trend for the fast PD readback for all 8 SOS optics (40 channels in all, I didn't look at the ETMs as their whitening boards are at the ends), and everything looks okay... This while situation seems very precarious to me, perhaps we should have a more robust signal routing from the OSEMs to the DAQ that is more immune to cable touching etc...

  12659   Fri Dec 2 16:21:12 2016 gautamUpdateGeneralrepaired projector, new mixer arrived and installed

The most recent power outage took out our projector and mixer. The projector was sent for repair while we ordered a new mixer. Both arrived today. Steve is working on re-installing the projector right now, and I installed the mixer which was verified to be working with our DAFI system (although the 60Hz issue still remains to be sorted out). The current channel configuration is:

Ch1: 3.5mm stereo output from pianosa

Ch2: DAFI (L)

Ch3: DAFI (R)

I've set some random gains for now, but we will have audio again when locking laugh

  12660   Fri Dec 2 16:40:29 2016 gautamUpdateIMC24V fuse pulled out

I've pulled out the 24V fuse block which supplies power to the AOM RF driver. The way things are set up on the PSL table, this same voltage source powers the RF amplifiers which amplify the green beatnote signals before sending them to the LSC rack. So I turned off the green beat PDs before pulling out the fuse. I then disconnected the input to the RF driver (it was plugged into a DS345 function generator on the PSL table) and terminated it with a 50 ohm terminator. I want to figure out a smart way of triggering the AOM drive and recording a ringdown on the scope, after which I will re-connect the RF driver to the DS345. The RF driver, as well as the green beat amplifiers and green beat PDs, remain unpowered for now...

  12663   Mon Dec 5 01:58:16 2016 gautamUpdateIMCIMC ringdowns

Over the weekend, I worked a bit on getting these ringdowns going. I will post a more detailed elog tomorrow but here is a quick summary of the changes I made hardware-wise in case anyone sees something unfamiliar in the lab...

  • PDA10CF PD installed on PSL table in the beam path that was previously used for the 3f cancellation trials
  • PDA255 installed on MC2 trans table, long BNC cable running from there to vertex via overhead cable tray
  • PDA255 installed on AS table in front of one of the (currently unused) WFS

I spent a while in preparation for these trials (details tomorrow) like optimizing AOM alignment/diffracted power ratio, checking AOM and PMC switching times etc, but once the hardware is laid out, it is easy to do a bunch of ringdowns in quick succession with an ethernet scope. Tonight I did about 12 ringdowns - but stupidly, for the first 10, I was only saving 1 channel from the oscilloscope instead of the 3 we want to apply the MIT method.

Here is a representative plot of the ringdown - at the moment, I don't have an explanation for the funky oscillations in the reflected PD signal, need to think on this.. More details + analysis to follow...


Dec 5 2016, 130pm:

Actually the plot I meant to put up is this one, which has the time window acquired slightly longer. The feature I am referring to is the 100kHz oscillation in the REFL signal. Any ideas as to what could be causing this?

Attachment 1: IMCringdown.pdf
IMCringdown.pdf
Attachment 2: IMCringdown_2.pdf
IMCringdown_2.pdf
  12664   Mon Dec 5 15:05:37 2016 gautamUpdateLSCMC1 glitches are back

For no apparent reason, the MC1 glitches are back. Nothing has been touched near the PD whitening chassis today, and the trend suggests the glitching started about 3 hours ago.. I had disabled the MC1 watchdog for a while to avoid the damping loop kicking the suspension around when these glitches occur, but have re-enabled it now. IMC is holding lock for some minutes... I was hoping to do another round of ringdowns tonight, but if this persists, its going to be difficult...

  12665   Mon Dec 5 15:55:25 2016 gautamUpdateIMCIMC ringdowns

As promised, here is the more detailed elog.


Part 1: AOM alignment and diffraction efficiency optimization

I started out by plugging in the input to the AOM driver back to the DS345 on the PSL table, after which I re-inserted the 24V fuse that was removed. I first wanted to optimize the AOM alignment and see how well we could cut the input power by driving the AOM. In order to investigate this, I closed the PMC, unlocked the PSL shutter, and dialed the PSL power down to ~100mW using the waveplate in front of the laser. Power before touching anything just before the AOM was 1.36W as measured with the Coherent power meter. 

The photodiode (PDA255) for this experiment was placed downstream of the 1%(?) transmissive optic that steers the beam into the PMC (this PD would also be used in Part 2, but has since been removed)...

Then I tuned the AOM alignment till I maximized the DC power on this newly installed PD. It would have been nicer to have the AOM installed on the mount such that the alignment screws were more easily accessible, but I opted against doing any major re-organization for the time being. Even after optimizing the AOM alignment, the diffraction efficiency was only ~15%, for 1V to the AOM driver input. So I decided to play with the AOM driver a bit.

Note that the AOM driver is powered by 24V DC, even though the spec sheet says it wants 28V. Also, the "ALC" input is left unconnected, which should be fine for our purposes. I opted to not mess with this for the time being - rather, I decided to tweak the RF adjust potentiometer on the front of the unit, which the spec sheet says can adjust the RF power between 1W and 2W. By iteratively tuning this pot and the AOM alignment, I was able to achieve a diffraction efficiency of ~87% (spec sheet tells us to expect 80%), in a switching time of ~130ns (spec sheet tells us to expect 200ns, but this is presumably a function of the beam size in the AOM). These numbers seemed reasonable to me, so I decided to push on. Note that I did not do a thorough check of the linearity of the AOM driver after touching the RF adjust potentiometer as Koji did - this would be relevant if we want to use the AOM as an ISS servo actuator, but for the ringdown, all that matters is the diffraction efficiency and switching time, which seemed satisfactory. 

At this point, I turned the PSL power back up (measured 1.36W just before the AOM). Before this, I estimated the PD would have ~10mW power incident on it, and I wanted it to be more like 1mW, so I I put an ND 1.0 filter on to avoid saturation.


Part 2: PMC "ringdown"

As mentioned in my earlier elog, we want the PMC to cut the light to the IMC in less than 1us. While I was at it, I decided to see if I could do a ringdown measurement for the PMC. For this, I placed two more PDs in addition to the one mentioned in Part 1. One monitored the transmitted intensity (PDA10CF, installed in the old 3f cancellation trial beam path, ~1mW incident on it when PMC is locked and well aligned). I also split off half the light to the PMC REFL CCD (2mW, so after splitting, PMC CCD gets 1mW through some ND filters, and my newly installed PD (PDA255) receives ~1mW). Unfortunately, the PMC ringdown attempts were not successful - the PMC remains locked even if we cut the incident light by 85%. I guess this isn't entirely surprising, given that we aren't completely extinguishing the input light - this document deals with this issue.... But the PMC transmitted intensity does fall in <200ns (see plot in earlier elog), which is what is critical for the IMC ringdown anyways. So I moved on.


Part 3: IMC ringdown

The PDA10CF installed in part 2 was left where it was. The reflected and transmitted light monitors were PDA255. The former was installed in front of the WFS2 QPD on the AS table (needed an ND1.0 filter to avoid damage if the IMC unlocks not as part of the ringdown, in which case ~6mW of power would be incident on this PD), while the latter was installed on the MC2 transmission table. We may have to remove the former, but I don't see any reason to remove the latter PD. I also ran a long cable from the MC2 trans table to the vertex area, which is where I am monitoring the various signals.

  

The triggering arrangement is shown below.

  

To actually do the ringdown, here is the set of steps I followed.

  1. Make sure settings on scope (X & Y scales, triggering) are optimized for data capture. All channels are set to 50ohm input impedance. The trigger comes from the "TTL" output of the DS345, whose "signal" output drives the AOM driver. Set the trigger to external, the mode should be "normal" and not "auto" (this keeps the data on the screen until the next trigger, allowing us to download the data via ethernet.
  2. The DS345 is set to output a low frequency (0.005Hz) square wave, with 1Vpp amplitude, 0.5V offset (so the AOM driver input is driven between 0V and 1V DC, which is what we want). This gives us ~100 seconds to re-lock the IMC, and download the data, all while chilling in the control room
  3. The autolocker was excellent yesterday, re-acquiring the IMC lock in ~30secs almost every time. But in the few instances it didn't work, turn the autolocker off (but make sure the MC2 tickle is on, it helps) and manually lock the IMC by twiddling the gain slider (basically manually do what the autolock script does). As mentioned above, you have ~100 secs to do this, if not just wait for 200secs and the next trigger...
  4. In the meantime, download the data (script details to follow). I've made a little wrapper script (/users/gautam/2016_12_IMCloss/grabChans.sh) which uses Tobin's original python script, which unfortunately only grabs data one channel at a time. The shell script just calls the function thrice, and needs two command line arguments, namely the base name for the files to which the data will be written, and an IP address for the scope...

It is possible to do ~15 ringdowns in an hour, provided the seismic activity is low and the IMC is in a good mood. Unfortunately, I messed up my data acquisiton yesterday, so I only have data from 2 ringdowns, which I will work on fitting and extracting a loss number from. The ringing in the REFL signal is also a mystery to me. I will try using another PDA255 and see if this persists. But anyways, I think we can exclude the later part of the REFL signal, and fit the early exponential decay, in the worst case. The ringdown signal plots have been uploaded to my previous elog. Also, the triggering arrangement can be optimized further, for example by using the binary output from one of our FEs to trigger the actual waveform instead of leaving it in this low frequency oscillation, but given our recent experience with the Binary Output cards, I thought this is unnecessary for the time being...

Data analysis to follow.


I have left all the PDs I put in for this measurement. If anyone needs to remove the one in front of WFS2, go ahead, but I think we can leave the one on the MC2 trans table there...

Attachment 2: AOMswitching.pdf
AOMswitching.pdf
Attachment 6: electricalLayout.pdf
electricalLayout.pdf
  12666   Mon Dec 5 19:29:52 2016 gautamUpdateIMCIMC ringdowns

The MC1 suspension troubles vanished as they came - but the IMC was remaining locked stably so I decided to do another round of ringdowns, and investigate this feature in the reflected light a bit more closely. Over 9 ringdowns, as seen in the below figure, the feature doesn't quite remain the same, but qualitatively the behaviour is similar.

Steve helped me find another PDA255 and so I will try switching out this detector and do another set of ringdowns later tonight. It just occurred to me that I should check the spectrum of the PD output out to high frequencies, but I doubt I will see anything interesting as the waveform looks clean (without oscillations) just before the trigger...

Attachment 1: REFLanomaly.pdf
REFLanomaly.pdf
  12667   Tue Dec 6 00:43:41 2016 gautamUpdateIMCmore IMC ringdowns

In an effor to see if I could narrow down the cause of the 100kHz ringing seen in the reflected PD signal, I tried a few things.

  1. Changed the PD - there was a PDA 255 sitting on the PSL table by the RefCav. Since it wasn't being used, I swapped the PD I was using with this. Unfortunately, this did not solve the problem.
  2. Used a different channel on the oscilloscope - ringing persisted
  3. Changed BNC cable running from PD to oscilloscope - ringing persisted
  4. Checked the spectrum of the PD under dark and steady illumination conditions for any features at 100kHz, saw nothing (as expected) 

I was working under the hypothesis that the ringing was due to some impedance mismatch between the PD output and the oscilloscope, and 4 above supports this. However, most documents I can find online, for example this one, recommend connecting the PD output via 50ohm BNC to a scope with input impedance 50ohms to avoid ringing, which is what I have done. But perhaps I am missing something.

Moreover, the ringdown in reflection actually supplies two of the five variables needed to apply the MIT method of loss estimation. I suppose we could fit the parameter "m4" from the ringdown in transmission, and then use this fitted value on the ringdown in reflection to see where the reflected power settles (i.e. the parameter "m3" as per the MIT paper). I will try analyzing the data on this basis.

I also measured the power levels at each of the PDs, these should allow us to calibrate the PD voltage outputs to power in Watts. All readings were taken with the Ophir power meter, with the filter removed, and the IMC locked.

PD Power level
REFL 0.47 mW (measured before 1.0 ND filter)
Trans 203 uW
Incident 1.06 mW

 

  12701   Tue Jan 10 22:55:43 2017 gautamUpdateCDSpower glitch - recovery steps

Here is a link to an elog with the steps I had to follow the last time there was a similar power glitch.

The RAID array restart was also done not too long ago, we should also do a data consistency check as detailed here, if not already..

If someone hasn't found the time to do this, I can take care of it tomorrow afternoon after I am back.

Quote:

Does "done" mean they are OK or they are somehow damaged? Do you mean the workstations or the front end machines?

The computers are all done.

megatron and optimus are not responding to ping commands or ssh -- please power them up if they are off; we need them to get data remotely

 

  12702   Wed Jan 11 16:35:03 2017 gautamUpdateCDSpower glitch - recovery progress

[lydia, ericq, gautam]

We set about following the instructions linked in the previous elog. A few notes/remarks:

  1. It is important to run the ntpdate commands before restarting the models. Sometimes, multiple restarts of the models were required to turn all the indicator blocks on the MEDM screen green.
  2. There was also an issue of multiple ntpd processes running on the same machine, which obviously caused all sorts of timing havoc. EricQ helped us diagnose and fix these. At the moment, all the lights are green on the CDS status MEDM screen
  3. On the hardware side, apart from the usual suspects of frontends/megatron/optimus/fb needing to be rebooted, I noticed that the ETMX OSEM lights were off on the control room monitors. Investigation pointed to the 2 20V sorensens at the X end outputting 0V, 0A after the power glitch. We turned down both dials, and then gradually ramped them up again. Both Sorensens now read +/-20V, 0.3A, which is in agreement with the label stuck onto them.
  4. Restarted MC autolocker and FSS Slow scripts on megatron. I have not yet looked at the status of the nds2 server on megatron.
  5. 11 MHz Marconi has yet to be restarted - but I am unable to get even the IMC locked at the moment. For some reason, the RMS of the MC1 and MC3 coils are way higher than what I am used to seeing (~5mV rms as compared to the <1mV rms I am used to seeing for a damped optic). I will investigate further. Leaving MC autolocker disabled for now.
  12708   Thu Jan 12 17:31:51 2017 gautamUpdateCDSDC errors

The IFO is more or less back to an operational state. Some details:

  1. The IMC mirror excess motion alluded to in the previous elog was due to some timing issues on c1sus. The "DAC" and "DK" blocks in the c1x02 diag word were red instead of green. Restarting all the models on c1sus fixed the problem
  2. When c1ioo was restarted, all of Koji's changes (digital) to the MC WFS servo where lost as they were not committed to the SDF. Eric suggested that I could just restore them from burt snapshots, which is what I did. I used the c1iooepics.snap file from 12:19PM PST on 26 December 2016, which was a time when the WFS servo was working well as per this elog by Koji. I have also committed all the changes to the SDF. IMC alignment has been stable for the last 4 hours.
  3. Johannes aligned and locked the arms today. There was a large DC offset on POX11, which was zeroed out by closing the PSL shutter and running LSC offsets. Both arms lock and stay aligned now.
  4. The doubling oven controller at the Y end was switched off. Johannes turned it on.
  5. Eric and I started a data consistency check on the RAID array yesterday, it has completed today and indicated no issues
  6. NDS2 is now running again on megatron so channel access from outside should(???) be possible again.

One error persists - the "DC" indicator (data concentrator?) on the CDS medm screen for the various models spontaneously go red and return to green often. Is this a known issue with an easy fix?

  12716   Fri Jan 13 23:39:46 2017 gautamUpdateGeneralETMX suspension electronics problems?

[Koji,gautam]

After Koji's leap second fix, we were playing around with the X arm locking. In particular, we were playing around with the limit value on the X arm LSC filter bank - the nominal value is 4000, we wanted to see if we could increase this without kicking the optic while acquiring arm lock. We initially increased it to 8000, and then turned it off altogether. Then we rapidly turned the output of the servo ON/OFF, and looked at the arm transmission to see if it came back to the level before unlocking, as an indication of whether the optic was kicked.

These trials suggested a value of 8000 for the limiter was OK, so we left the LSC mode on with the limiter set to 8000. But just as we were about to leave for the night, I noticed on the wall Striptool that the X arm was unlocked. Investigating, we found that the green wasn't even locking to a HOM. Further investigation of the Oplev spot showed that ETMX had received a large kick (both pitch and law errors were ~200urad). ITMX was unaffected.

We initially tried lowering the LSC limit value back to 4000, then used first the Oplev spot and then the green to align the arm. But turning on LSC misaligned the arm after acquiring lock. So we decided to leave LSC off, thinking that the notorious ETMX suspension problems have resurfaced. As a diagnostic, we figured we'd leave the watchdog tripped, and use the Oplev to see if the optic was getting kicked. But the act of turning the watchdog off kicked the optic again (WHY?!).

Looking at the ETMX sus screen, turning off all the damping and LSC (but watchdog on) still leaves a non-zero offset in the "Vmon" field, between 0.02-0.05V depending on the coil. Turning the watchdog OFF takes all these to 0.009V, although I can see the LR value fluctuating between 0.004V and 0.009V. I went to the Xend and squished all the cables on the Sat. Box, but the problem persisted.

At this time, I can't think of any explanation, so I am giving up for the night. To avoid unnecessarily kicking the optic, I am going to unplug the suspension from the Sat. Box and leave one of our tester boxes plugged in, lets see if that sheds any light on the situation...


Notes:

  1. The +/-20V sorensens at this end were "tripped" for a few days after the power glitch until they were reset and turned back on yesterday. But this should not affect Vmon, as these Sorensens only supply the DC voltage for the coil bias, which is a slow machine channel?
  2. The X arm was staying locked and well aligned for hours on end earlier this afternon - in fact it was locked for about 2 hours 6-8 hours ago, I can still see the trace on the wall StripTool....
  12725   Mon Jan 16 23:25:07 2017 gautamUpdateSUSMC1 SUS electronics investigation

[rana,gautam]

Summary:

  • MC1 glitchy behaviour is back
  • Found a broken LEMO cable, left unplugged for the night -> to be repaired tomorrow
  • Further diagnosis to follow

During the course of Rana's inspection of the general state of the IFO, he commented that there seemed to be several seismic-related IMC lock losses in the time that he had been observing it. This issue looked suspiciously like the the MC1 glitches I had noticed sometime late last year, especially since each time the IMC would unlock, we could see significant amounts of motion on MC REFL. To diagnose, we did the following:

  1. Closed PSL shutter
  2. Ramped down the gains of the MC1 damping loops by factor of 1000 in ~4 secs using z step
  3. Shut down the watchdog for MC1
  4. Observed dataviewer traces for glitches

Sure enough, there were several glitches that occurred in all 5 sensor channels. These glitches varied in size from a few counts (the smaller ones) to 60-70counts for the bigger ones. In the past, squishing the LEMO connector on the front of the PD whitening board (D000210) had apparently made the glitching go away. So tonight, for starters, we squished everything else - Sat. Box connectors, the breakout board from Sat. Box to whitening board on the back of 1X6, and the DB connector on the front of the whitening board. This had no effect - the glitching remained consistent.

Next, Rana pulled out two of the three 4pin LEMOs, and left only those coresponding to UL/LL plugged in - but the glitching persisted in these two channels. We then pulled out the board. It was installed in 1998, but has a sticker on it that says "fixed in 2003". Not sure what the fix was. Visual inspection of the circuit didn't show anything obviously faulty, but it did look like the two MAX333A quad switches (these control whether the whitening is bypassed or not) had been replaced at some point. There are other undesirable features, such as the use of thick film resistors, but nothing that would explain the glitchy behaviour.

Next, we re-inserted the whitening board back into its original slot in the Eurocrate, but switched the cables (both D sub and LEMO, but only on the whitening board end) between the boards for MC1 and MC3 (i.e. MC1 cables were routed through the whitening board that was originally used for MC3, and vice-versa). But the glitches remained consistent on the MC1 channels. So it looks like the board is not a likely culprit.

Finally, we went in and squished all the cables from the PD whitening board to the ADC (via an AA filter board). For some of the LEMO cables from the whitening board, the LEMO backshells were not properly tightened. Rana fixed these before putting them back in. Some of the connectors were also not pushed in tightly enough, Rana heard the click when he pushed them in. The cables from the adaptor board to the ADC itself looked fine, it was screwed on at both ends, and all these connections looked snug enough. In the interest of completeness, Rana also pushed in the backplane connectors on the Eurocrate (these supply the signals from the BIO cards to switch the whitening ON/OFF). The one corresponding to MC1 was indeed a little loose.

Coming back to the control room, we saw that the MC1 LR sensor was dead. After some investigation, Rana found that on the AA filter board end, one of the 4pin LEMOs from the whitening board had one of its wires come unstuck from where it was soldered (this presumably happened while we were squishing cables tonight, as the LR channel was fine before that). Also, there was no heat shrink used on any of the solder joints. Could this explain the glitchy behaviour? Perhaps, but the glitches remained in the 3 channels that were connected. Anyways, I will repair this cable tomorrow, and we can see if this has fixed the problem or not..


Some misc points:

  1. Regarding the adaptor boards that take the PD signals from the satellite box and route it to the whitening board, there are some clamps that hold the IDE connectors in place for MC1, MC2 and MC3 boards, but not for the others (see attached picture). Steve, can we install clamps for all of the boards? [taken care of, see here]
  2. The whitening boards are not screwed in place into the Eurocrate. This should be rectified.

PSL shutter is closed, MC1 watchdog is shutdown for the night.

Attachment 1: 20170116_231625.png
20170116_231625.png
Attachment 2: IMG_7175.JPG
IMG_7175.JPG
Attachment 3: IMG_7174.JPG
IMG_7174.JPG
  12728   Tue Jan 17 21:29:52 2017 gautamUpdateSUSMC1 SUS electronics investigation

 

Quote:
 

After some investigation, Rana found that on the AA filter board end, one of the 4pin LEMOs from the whitening board had one of its wires come unstuck from where it was soldered (this presumably happened while we were squishing cables tonight, as the LR channel was fine before that). Also, there was no heat shrink used on any of the solder joints.

The faulty cable has been re-soldered (with heat shrink) and replaced. All 5 sensor signals appear normal on dataviewer now. I am leaving things in this state for the night, let us see if the glitches return overnight.

PSL shutter remains closed

  12729   Tue Jan 17 21:31:57 2017 gautamUpdateGeneralETMX suspension electronics problems?

Last night, I plugged the ETMX suspension coils back into the satellite box. Tonight, we turned on the damping loops for ETMX. Rana centered the Oplev so we can use that as an additional diagnostic to see if the optic gets kicked around overnight. We will re-assess the situation tomorrow.

Sometime earlier today, Lydia noticed that the +/- 5V Sorensens at the X end were not displaying their nominal voltage/current values (as per the stickers on them). She corrected this.

  12730   Wed Jan 18 10:41:14 2017 gautamUpdateGeneralETMX suspension electronics problems?

Summary pages show no kicking in the ETMX watchdogs from midnight to 6 AM (0800 - 1400 UTC):

https://nodus.ligo.caltech.edu:30889/detcharsummary/day/20170118/sus/watchdogs/

  12731   Wed Jan 18 11:40:54 2017 gautamUpdateSUSMC1 SUS electronics investigation

After the repair of the faulty LEMO cable, I left MC1 with it's watchdog off overnight. Unfortunately, it looks like the problem still persists. The first attachment shows a second trend plot for the past 15 hours. Towards the left end of the plot, you can see where I re-connected the LEMO cable for the LR/UR channels.

A couple of months ago, I added a BLRMS block for the IMC optics that calculates BLRMS for the shadow sensor output as well as the coil output. Looking at this trend overnight, I noticed that the glitches appear in the coil outputs as well, as shown in the plot below, which is for a 1 hour stretch last night (I used the full data from a 16Hz coil output channel and not the BLRMS, I am not sure if there is a DQ'ed version of the coil outputs?).

Zooming in further to one of these glitches, we can see that the glitches in the coil and shadow sensor signals are in fact coincident.

But given that the watchdog was turned off all this time, the only voltage going to the coils should be the DC bias voltages. So does this not support the hypothesis that the problem lies in the part of the signal chain that supplies the bias voltage to the coils?

Never mind, the "coil output" channel isn't a true readback of the voltage to the coil, but is the calculated damping output (which is not sent to the coils when the watchdog is shutdown...

  12734   Wed Jan 18 14:23:47 2017 gautamUpdateSUSMC1 SUS electronics investigation

As part of the ongoing debugging, I've switched the MC1 and MC3 satellite boxes. Both MC1 and MC3 have their watchdogs shudown for the moment.

  12736   Wed Jan 18 18:44:53 2017 gautamUpdateSUSMC1 SUS electronics investigation
Quote:

As part of the ongoing debugging, I've switched the MC1 and MC3 satellite boxes. Both MC1 and MC3 have their watchdogs shudown for the moment.

In the last 3.5 hours, there has been nothing conclusive - no evidence of any glitching in either MC1 or MC3 sensor channels. I am going to hold off on doing the LEMO cable swap test for a few more hours, to see if we can rule out the satellite box.

  12739   Thu Jan 19 12:00:10 2017 gautamUpdateSUSMC1 SUS electronics investigation

Going through the last ~20 hours of data, the MC1 sensor channels look glitch free the entire period. However, there is a ~10min period around 1PM UTC today when there were a couple of glitches ~80 counts in size in all the MC3 sensor channels. The attached shows the full 2k data from all 10 channels (MC1 and MC3 sensors) around this time.

Is this sufficient evidence to conclude that the Satellite boxes are to blame? It's hard to explain why the glitches come and go in this fashion, and also the apparent difference in the length of time for which the glitches persist. Here, in almost 24 hours, there is one incidence of glitching, but in yesterday's trend plot, the glitching remains present over several hours... The amplitude of the glitches, and their coincidence in all 5 channels, seems consistent with what we have been seeing though...

 

  12742   Fri Jan 20 11:16:30 2017 gautamUpdateSUSMC1 SUS electronics investigation

Both suspensions have been relatively well behaved for the best part of the last two days, since I effected the Satellite Box swap. Today morning, I set about re-enabling the damping and locking the MC. Judging by the wall StripTool, it stayed locked for about 30 mins or so, after which the glitching returned.

Attached is a screenshot of the sensor signals from MC1 and MC3 (second trend), and also the highest band (>30Hz) BLRMS output for the same 10 channels (full data sampled at 16Hz). Note that MC1 and MC3 satellite boxes remain swapped. So the glitches now have migrated to the MC3 channels.

I need to think about whether this is just coincidence, or if me re-enabling the damping has something to do with the re-occurrence of the glitching...


Addendum 4.30pm: I've also re-aligned the Y arm. Its alignment has been stable over the last few hours, despite several mode cleaner lock losses in between, it recovers good IR transmission. The X arm has been re-aligned to green, but I can't get it locked to the IR - everytime I turn the LSC to ETMX on, there seems to be some large misalignment applied to it. c1iscaux was dead, I restarted it by keying the crate. I haven't had time to investigate the X arm locking in detail, I will continue to debug this.

  12746   Mon Jan 23 15:16:52 2017 gautamUpdateOptical LeversETMY Oplev HeNe needs to be replaced

On the control room monitors, I noticed that the IR TEM00 spot was moving around rather a lot in the Y arm. The last time this happened had something to do with the ETMY Oplev, so I took a look at the 30 day trend of the QPD sum, and saw that it was decaying steeply (Steve will update with a long term trend plot shortly). I noticed the RIN also seemed rather high, judging by how much the EPICS channel reading for the QPD sum was jumping around. Attached are the RIN spectra, taken with the OL spot well centered on the QPD and the arms locked to IR. Steve will swap the laser out if it is indeed the cluprit.

Attachment 1: ETMY_Oplev.pdf
ETMY_Oplev.pdf
  12748   Tue Jan 24 01:04:16 2017 gautamSummaryIOOIMC WFS RF power levels

Summary:

I got around to doing this measurement today, using a minicircuits bi-directional coupler (ZFBDC20-61-HP-S+), along with some SMA-LEMO cables.

  • With the IMC "well aligned" (MC transmission maximized, WFS control signals ~0), the RF power per quadrant into the Demod board is of the order of tens of pW up to a 100pW.
  • With MC1 misaligned such that the MC transmission dropped by ~10%, the power per quadrant into the demod board is of the order of hundreds of pW.
  • In both cases, the peak at 29.5MHz was well above the analyzer noise floor (>20dB for the smaller RF signals), which was all that was visible in the 1MHz span centered around 29.5 MHz (except for the side-lobes described later).
  • There is anomalously large reflection from Quadrant 2 input to the Demod board for both WFS
  • The LO levels are ~-12dBm, ~2dBm lower than the 10dBm that I gather is the recommended level from the AD831 datasheet
Quote:

We should insert a bi-directional coupler (if we can find some LEMO to SMA converters) and find out how much actual RF is getting into the demod board.


Details:

I first aligned the mode cleaner, and offloaded the DC offsets from the WFS servos.

The bi-directional coupler has 4 ports: Input, Output, Coupled forward RF and Coupled Reverse RF. I connected the LEMO going to the input of the Demod board to the Input, and connected the output of the coupler to the Demod board (via some SMA-LEMO adaptor cables). The two (20dB) coupled ports were connected to the Agilent spectrum analyzer, which have input impedance 50ohms and hence should be impedance matched to the coupled outputs. I set the analyzer to span 1MHz (29-30MHz), IF BW 30Hz, 0dB input attenuation. It was not necessary to turn on averaging to resolve the peaks at ~29.5MHz since the IF bandwidth was fine enough.

I took two sets of measurements, one with the IMC well aligned (I maximized the MC Trans as best as I could to ~15,000 cts), and one with a macroscopic misalignment to MC1 such that the MC Trans fell to 90% of its usual value (~13,500 cts). The peak function on the analyzer was used to read off the peak height in dBm. I then converted this to RF power, which is summarized in the table below. I did not account for the main line loss of the coupler, but according to the datasheet, the maximum value is 0.25dB so there numbers should be accurate to ~10% (so I'm really quoting more S.Fs than I should be).

WFS Quadrant Pin (pW) Preflected(pW) Pin-demod board (pW)

IMC well aligned

1 1 50.1 12.6 37.5
2 20.0 199.5 -179.6
3 28.2 10.0 18.2
4 70.8 5.0

65.8

2 5 100 19.6 80.0
6 56.2 158.5 -102.3
7 125.9 6.3 11.5
8 17.8 6.3

119.6
 

WFS Quadrant Pin (pW) Preflected(pW) Pin-demod board (pW)

MC1 Misaligned

1 1 501.2 5.0 496.2
2 630.6 208.9 422
3 871.0 5.0 866
4 407.4 16.6

190.8

2 5 407.4 28.2 379.2
6 316.2 141.3 175.0
7 199.5 15.8 183.7
8 446.7 10.0 436.7

 

For the well aligned measurement, there was ~0.4mW incident on WFS1, and ~0.3mW incident on WFS2 (measured with Ophir power meter, filter out).

I am not sure how to interpret the numbers for quadrants #2 and #6 in the first table, where the reverse coupled RF power was greater than the forward coupled RF power. But this measurement was repeatable, and even in the second table, the reverse coupled power from these quadrants are more than 10x the other quadrants. The peaks were also well above (>10dBm) the analyzer noise floor 

I haven't gone through the full misalginment -> Power coupled to TEM10 mode algebra to see if these numbers make sense, but assuming a photodetector responsivity of 0.8A/W, the product (P1P2) of the powers of the beating modes works out to ~tens of pW (for the IMC well aligned case), which seems reasonable as something like P1~10uW, P2 ~ 5uW would lead to P1P2~50pW. This discussion was based on me wrongly looking at numbers for the aLIGO WFS heads, and Koji pointed out that we have a much older generation here. I will try and find numbers for the version we have and update this discussion.

Misc:

  1. For the sake of completeness, the LO levels are ~ -12.1dBm for both WFS demod boards (reflected coupling was negligible)
  2. In the input signal coupled spectrum, there were side lobes (about 10dB lower than the central peak) at 29.44875 MHz and 29.52125 MHz (central peak at 29.485MHz) for all of the quadrants. These were not seen for the LO spectra.
  3. Attached is a plot of the OSEM sensor signals during the time I misaligned MC1 (in both pitch and yaw approximately by equal amounts). Assuming 2V/mm for the OSEM calibration, the approximate misalignment was by ~10urad in each direction.
  4. No IMC suspension glitching the whole time I was working today yes

 

Attachment 1: MC1_misalignment.png
MC1_misalignment.png
  12751   Wed Jan 25 01:27:45 2017 gautamUpdateIMCIMC feedforward checkup

This is probably just a confirmation of something we discussed a couple of weeks back, but I wanted to get more familiar with using the multi-coherence (using EricQs nice function from the pynoisesub package) as an indicator of how much feedforward noise cancellation can be achieved. In particular, in light of our newly improved WFS demod/whitening boards, I wanted to see if there was anything to be gained by adding the WFS to our current MCL feedforward topology.

I used a 1 hour data segment - the channels I looked at were the vertex seismometer (X,Y,Z) and the pitch and yaw signals of the two WFS, and the coherence of the uncorrelated part of these multiple witnesses with MCL. I tried a few combinations to see what is the theoretical best achievable subtraction:

  1. Vertex seismometer X and Y channels - in the plot, this is "Seis only"
  2. Seis + WFS 1 P & Y
  3. Seis + WFS 2 P & Y
  4. Seis + WFS 1 & 2 P
  5. Seis + WFS 1 & 2 Y

The attached plot suggests that there is negligible benefit from adding the WFS in any combination to the MCL feedforward, at least from the point of view of theoretical achievable subtraction

I also wanted to put up a plot of the current FF filter performance, for which I collected 1 hour of data tonight with the FF on. While the feedforward does improve the MCL spectrum, I expected better performance judging by previous entries in the elog, which suggest that the FIR implementation almost saturates the achievable lower bound. The performance seems to have degraded particularly around 3Hz, despite the multi-coherence being near unity at these frequencies. Perhaps it is time to retrain the Weiner filter? I will also look into installation of the accelerometers on the MC2 chamber, which we have been wanting to do for a while now...

Attachment 1: IMC_FF_potential.pdf
IMC_FF_potential.pdf
  12754   Wed Jan 25 14:30:20 2017 gautamUpdateCDSslow machine bootfest

[gautam, lydia]

We rebooted c1psl, c1iscaux and c1aux which were all showing the typical symptom of responding to ping but not to telnet (and also blanked out epics fields on the MEDM screens). Keyed all these crates.

Restored burt snapshots for c1psl, PMC locked fine, and IMC is also locked now.

Johannes forgot to elog this yesterday, but he rebooted c1susaux following the usual procedure to avoid getting ITMX stuck. 

 

  12759   Fri Jan 27 00:14:02 2017 gautamSummaryIOOIMC WFS RF power levels

It was raised at the Wednesday meeting that I did not check the RF pickup levels while measuring the RF error signal levels into the Demod board. So I closed the PSL shutter, and re-did the measurement with the same measurement scheme. The detailed power levels (with no light incident on the WFS, so all RF pickup) is reported in the table below.

IMC WFS RF Pickup levels @ 29.5MHz
WFS Quadrant Pin (pW) Preflected
1 1 0.21 10.
2 1.41 148
3 0.71 7.1
4 0.16 3.6
2 1 0.16 10.5
2 1.48 166
3 0.81 5.1
4 0.56 0.33

These numbers can be subtracted from the corresponding columns in the previous elog to get a more accurate estimate of the true RF error signal levels. Note that the abnormal behaviour of Quadrant #2 on both WFS demod boards persists.

  12765   Fri Jan 27 20:52:36 2017 gautamUpdateCDStest of new daqd code on fb1

I'm not sure if this is related, but since today morning, I've noticed that the data concentrator errors have returned. Looking at daqd.log, there is a 1 second timing mismatch error that is being generated. Usually, manually running ntpdate on the front ends fixes this problem, but it did not work today.

Attachment 1: DCerrors.png
DCerrors.png
  12766   Fri Jan 27 21:21:35 2017 gautamUpdateCDSc1pem revamped

The coil and PD BLRMS are useful tools in identifying when glitches occur in the PD  readout, I thought it would be good to install them for ITMY, ETMX and SRM (since I plan to switch the MC3 satellite box, which we suspect to be problematic, with the SRM one). For this purpose, I had to install some IPC SHMEM blocks in C1SUS and recompile. 24 IPC channels were added to pipe the coil, PD and Oplev signals from C1SUS to C1PEM - the recompilation went smoothly, and it doesn't look like the model computation time has increased significantly or that the model is any closer to timing out.

However, I was unable to install the BLRMS blocks in C1PEM, as when I tried to compile the model with BLRMS for these extra 24 channels, I got a compilation error saying that I have exceeded the maximum allowed 499 testpoints per channel. Is there any workaround to this? It would be possible to create a custom BLRMS block that doesn't have all those testpoints, maybe this is the way to go? Especially if we want to install these channels for all our SOS optics, and also replace the current Seismic BLRMS with this scheme for consistency?

GV edit: I have implemented this scheme - after backing up the original BLRMS_2k part, I made a new one with no testpoints and only EPICS readouts. Doing so allowed me to recompile c1pem without any issues, the CPU time seems to have gone up by 3us from ~55us to 58us. So the BLRMS data record is only available at 16Hz, since there are no DQ channels in the BRLMS block - do we want these in any case? Let's see how this does over the weekend...

  12768   Sat Jan 28 01:25:51 2017 gautamUpdateIMC29.5 MHz modulation depth

Some more details of our investigation:

  1. Here is a spectrum of the signal to the power combiner on the PSL table, measured on the output of the RF AM Stabilization box.

    Perhaps these sidebands were the ones I observed while looking at the input to the WFS demod board.
  2. The signal looked like a clean sinusoid when viewed on an oscilloscope with input impedance set to 50ohms. There were no sharp features or glitches in the time we observed, except when the 29.5 MHz MEDM slider was increased beyond 5, as noted by Lydia.
  3. We couldn't find a schematic for this RF AM Stabilization servo, so we are not sure what RF output power to the EOM we should expect. Schematic has since been found.
  4. I measured the power level at the input side (i.e. from the crystal) and found that it is ~12dBm, which seems reasonable (the front panel of the box housing the 29.5 MHz oscillator is labelled 13dBm). The schematic for the RF AM stabilization box says we should expect +10dBm at the input side, so all this points to a problem in the RF AM stabilization circuit...
  5. There is an attenuator dial on the front panel of the said RF AM stabilization servo that allows one to tune the power to the LO input of the WFS. Right now, it is set to approximately 7dB of attentuation, which corresponds to -12dBm at the WFS demod board input. I did a quick check to see if turning the dial changed the signal level at the LO input of the WFS board. The dial moves in clicks of 1dB, and the RF power at the LO input of the demod board increased/decreased by ~1dBm for each click the dial was rotated (I only explored the region 3dB-11dB of atttentuation). So it should be possible to increase the LO level to the WFS demod boards, is there any reason we shouldn't increase this to -8bBm (~0.25Vpp into 50ohms, which is around the level Koji verified the mixer to be working well at)?
  6. There were a couple of short ribbon cables which were just lying around on top of the cards in the eurocrate, Koji tells me that these were used as tester cables for checking the whitening filters and that they don't serve any purpose now. These have been removed.
  7. Added a button to IMC MEDM screen to allow easy access to the MEDM screen with slider to control the 29.5MHz modulation depth - though as mentioned in Lydia's elog, at the moment, this slider has no effect on the 29.5MHz power level to the EOM...
Attachment 1: IMC_mod.pdf
IMC_mod.pdf
  12771   Mon Jan 30 19:07:48 2017 gautamUpdateIMCRF AM stabilization box pulled out

[johannes, gautam]

We pulled out the RF AM stabilization box from the 1X2 rack. PSL shutter was closed, marconi output, RF distribution box and RF AM stabilization box were turned off in that order. We had to remove the 4 rack nut screws on the RF distribution box because of the stiff cables which prevented the RF AM stabilization box extraction. I've left the marconi output and the RF distribution boxes off, and have terminated all open SMA connections with 50 ohm terminators just in case. Rack nuts for RF distribution box have been removed, it is currently sitting on a metal plate that is itself screwed onto the rack. I deemed this a stable enough ledge for the box to sit on in the short run, while we debug the RF AM stabilization box. We will work on the debugging and re-install the box as soon as we are done...

  12775   Tue Jan 31 14:17:48 2017 gautamUpdateIMCRF AM stabilization box pulled out

> What is the probe situation? Ought to use a high impedance FET probe to measure this or else the scope would load the circuit.

We did indeed use the active probe, with the 100:1 attenuator in place. The values Lydia has quoted have 40dB added to account for this.

> What kind of HELA are the HELA amplifiers? Please a link to the data sheet if you can find it. I wonder what the gain and NF are at 30 MHz. I think the HELA-10D should be a good variant

The HELA is marked as HELA-10. It doesn't have the '+' suffix but according to the datasheet, it seems like it is just not RoHS compliant. It isn't indicated which of the varieties (A-D) is used either on the schematic or the IC, only B and D are 50ohms. For all of them, the typical gain is 11-12dB, and NF of 3.5dB.

  12778   Tue Jan 31 18:51:07 2017 gautamUpdateSEISeismic Rainbow Strip - myths debunked

I've been suggesting that there may be something wonky with the Seismic Rainbow Striptool on the wall for the last couple of weeks. Here are a few things that were verified today.

  1. If you want to restore the StripTools in the control room, just run /opt/rtcds/caltech/c1/scripts/general/startStrip.sh. I have verified as of today that this works, and in future, any changes to channels/limits/colors of traces etc should be reflected in this script.
  2. Though some of the BLRMS bands have looked anomalous over the last few weeks, in particular the 0.3-1Hz band. The attached 120 day trend plot suggests that there hasn't been any dramatic change recently. In fact, looking on the summary pages, Rana noticed that today was an unusually low 0.3-1Hz activity day..
Attachment 1: Seis_BLRMS.png
Seis_BLRMS.png
  12780   Tue Jan 31 22:07:13 2017 gautamUpdateIMCRF AM stabilization box revamp

I've added the schematic of the RF AM stabilization board to the 40m PSL document tree, after having created a new DCC document for our 40m edits. Pictures of the board before and after modification will also be uploaded here...

  12790   Thu Feb 2 17:43:20 2017 gautamUpdatePEMEM172 mic is hooked up in the PSL

I had noticed something wonky with the microphone, but neglected to elog it. I had tested it after installation by playing a sine wave from my laptop and looking at the signal on the PSL table, it worked fine. But you can see in the attached minute trend plot that the signal characteristics changed abruptly ~half a day after installation, and never quite recovered.\

 

Attachment 1: Mic_broken.png
Mic_broken.png
  12793   Fri Feb 3 00:36:52 2017 gautamUpdateIMCMCL Feedback - framing the problem

Rana motivated me to take a step back and reframe the objectives and approach for this project, so I am collecting some thoughts here on my understanding of it. As I write this, some things still remain unclear to me, so I am leaving these as questions here for me to think about...

Objectives

  1. The PSL is locked to the IMC cavity - but at frequencies near 1 Hz, the laser frequency is forced to follow the IMC cavity length fluctuations, even though the free-running PSL frequency noise at those frequencies is lower. This excess is also imprinted on the arms when locked to the IR. We would like to improve the situation by feeding back a portion of the MC PDH error signal to the cavity length actuator to stabilize the MC cavity length at low frequencies. Moreover, we would like this loop to not imprint additional control noise in the arm control signals, which is a problem we have observed with the existing MCL loop. 
     
  2. The borader goal here is to use this project as a case study for designing the optimal loop and adaptive feedback. Can we come up with an algorithm, which takes
    • A model of our system (made with measured data where possible)
    • A list of our requirements (e.g. in this case, frequency noise requirements in various frequency bands, smooth crossovers between the various loops that enable locking the PSL to the IMC cavity and avoid injecting excess control noise into the plant)

and come up with the best loop that meets all our rquirements? What constitutes the "best" loop? How do we weight the relative importance of our various requirements? 


Proposed approach:

For the specific problem of making the MCL feedback loop better, the approach I have in mind right now is the following:

  1. Build a model of the 40m IMC loop. Ultimately the performance of the loop we implement will depend on the transfer function from various additive noise sources and disturbances in the feedback loop (e.g. electronics noise) to the output (i.e. laser frequency). Building an accurate model will allow us to quantify the performance of the proposed control loop, and hence, optimize it with some algorithm. I did some work on a simplistic, purely analytical model of the two MC loops (MCF and MCL), but Rana pointed out that it is better to have something more realistic for this purpose. I have inherited his Simulink models, which I will now adapt to reflect the 40m topology. 
  2. Come up with a list of requirements for the MCL controller. Some things that come to mind:
    • Reduce the arm control signal spectral amplitude below 20 Hz
    • Not increase the arm control signal spectral amplitude above 20 Hz
    • Crossover smoothly with the FSS slow temperature control loop and the MCF loop. 
    • What factor of suppression are we looking for? What is achievable? Once I build the model, it should shed some light on these..
    • Is the PMC a more stable frequency reference than the NPRO crystal at low frequencies? This measurement by Koji seems to suggest that it isn't (assuming the 1e4 product for the NPRO free-running frequency noise)..
  3. Once we have a model and a satisfactory list of requirements, design a control loop that meets these using traditional techniques, i.e. desired tracking error in the control band of 0.1-20 Hz (is this possible? The model will tell us...), gain and phase margin requirements etc. But this need not necessarily be the optimal controller that meets all of our requirements
  4. Optimize the controller - how? Can we define an objective function that, for example, rewards arm control signal suppression and penalizes injection of control noise, and just fminsearch in the [z,p,k] parameter space of the controller? Is there a smarter way to do this?
  5. Can this algorithm be adaptive, and optimize the controller to adapt to prevailing seismic conditions for example? Is this the same as saying we have a model that is accurate enough for us to predict the response of the plant to environmental disturbances? 

My immediate goal is to have the Simulink model updated.

Thoughts/comments on the above will be appreciated...

 
  12803   Mon Feb 6 15:18:08 2017 gautamUpdateCDSslow machine bootfest

Had to reboot c1psl, c1susaux, c1auxex, c1auxey and c1iscaux today. PMC has been relocked. ITMX didn't get stuck. According to this thread, there have been two instances in the last 10 days in which c1psl and c1susaux have failed. Since we seem to be doing this often lately, I've made a little script that uses the netcat utility to check which slow machines respond to telnet, it is located at /opt/rtcds/caltech/c1/scripts/cds/testSlowMachines.bash.

The script can be executed by ./testSlowMachines.bash.

  12804   Mon Feb 6 17:03:41 2017 gautamUpdateIMCMCL Feedback - simulink model updated

I've edited Rana's Simulink model to reflect the current IMC servo topology (to the best of my understanding). I've tried to use Transfer Function blocks wherever possible so that we can just put in the appropriate zpk model in the script that will linearize the whole loop. I've also omitted the FSS SLOW loop for now.

I've been looking through some old elogs and it looks like there have been several modifications to both the MC servo board (D040180) and the TT FSS Box (D040105). I think it is easiest just to measure these TFs since the IMC is still down, so I will set about doing that today. There is also a Pomona Box between the broadband EOM and the output of the TT FSS box, which is meant to sum in the modulation for PMC locking, about which I have not yet found anything on the elog.

So the next steps are:

  1. Measure/estimate all the unknown TFs and gains in this schematic
  2. Linearize the model, get the OLG, see if the model matches previously measured OLGs (with the MCL part disabled)
  3. Once the model is verified to be correct, look at couplings of various noise sources in the MCL part of the loop, and come up with a suitable controller.

If anyone sees something wrong with this topology, please let me know so that I can make the required changes.

Attachment 1: mc40_v1.pdf
mc40_v1.pdf
  12806   Tue Feb 7 10:18:58 2017 gautamUpdateIMCMC REFL weirdness

A few minutes back, I glanced up at the control room StripTool and noticed that the MCREFL PD DC level had gone up from ~0 to ~0.7, even though the PSL shutter was closed. This seemed bizzare to me. Strangely, simply cycling the shutter returned the value to the expected value of 0. I wonder if this is just a CDS problem to do with c1iool0 or c1psl? (both seem to be responding to telnet though...)

Since things look to be back to normal, I am going to start with my characterization of the various TFs in the IMC FSS loop...

  12812   Wed Feb 8 19:13:02 2017 gautamUpdateIMCMCL Feedback - TF measurements

Quick summary elog, details to follow. I did the following:

  • Updated the Simulink model based on Koji's feedback. 
  • Today morning, I measured the (electronic) open-loop TFs of
    • MC Servo Board
    • FSS Fast path (PZT)
    • FSS PC Drive path
  • The summing amplifiers in the latter two paths are assumed to be broadband for the purposes of this model.

The measurements I have look reasonable. But I had a hard time trying to look at the schematic and determine what is the appropriate number and locations of poles/zeros with which to fit the measured transfer function. Koji and I spent some time trying to go through the MC Servo board schematic, but looks like the version uploaded on the 40m DCC tree doesn't have changes made to it reflected (we compared to pictures on the 40m google photos page and saw a number of component values were different). Since the deviation between fit and measurement only occurs above 1MHz (while using poles/zeros inferred from the schematic), we decided against pulling out the servo board and investigating further - but this should be done at the next opportunity. I've marked the changes we caught on a schematic and will upload it to the 40m DCC page, and we can update this when we get the chance.

So it remains to fit the other two measured TFs, and add them to the Simulink model. Then the only unknown will be the PDH discriminant, which we anyway want to characterize given that we will soon have much more modulation.  

Data + plots + fits + updated schematics to follow...

 

  12814   Thu Feb 9 11:22:56 2017 gautamUpdateGeneralSorensens and DIN connections at 1X1

I'd like to fix a few things at 1X1 when we plug in the new amplifier for the 29.5MHz modulation signal. 

  1. Split off separate +24 and ground wires to the green BBPD RF amplifiers and the AOM driver (they are sharing a single fuse at the moment)
  2. Tap a new +24 GND -24V set for the FSS Fast summing box - this is currently running with a bench power supply underneath the PSL table set to +/-18V, but I checked the 7815/7915 datasheets and they accept up to 35V input for a 15V output, so it should be fine to use 24V
  3. Hook up the ZHL-2A for the IMC modulation.

Steve has ordered rolls of pre-twisted wire to run from 1X1 to the PSL table, so that part can be handled later.

But at 1X1, we need to tap new paths from +/- 24V to the DIN connectors. I think it's probably fine to turn off the two Sorensens, do the wiring, and then turn them back on, but is there any procedure for how this should be done? 

Attachment 1: Screen_Shot_2017-02-10_at_9.01.46_AM.png
Screen_Shot_2017-02-10_at_9.01.46_AM.png
  12815   Thu Feb 9 23:35:34 2017 gautamUpdateIMCMCL Feedback - TF measurements

Here are the details as promised.

Attachment #1: Updated simulink model. Since I haven't actually run this model, all the TF blocks are annotated "???", but I will post an updated version once I have run the model (and fix some of the questionable aesthetic choices)

Attachment #2: Measured and fitted transfer functions from the "IN1" input (where the demodulated MC REFL goes) to the "SERVO" output of the MC servo board (to FSS box). As mentioned in my previous elog, I had to put in a pole (fitted to be at ~2MHz, called pole 9 in the plot) in order to get good agreement between fit an measurement up to 10MHz. I didn't bother fitting all the high frequency features. Both gain sliders on the MEDM screen ("IN1 Gain" and "VCO gain") were set to 0dB for this measurement, while the super boosts were all OFF.

Attachment #3: Measured and fitted transfer function from "TEST 1 IN" to "FAST OUT" of the FSS box. Both gains on the FSS MEDM screen ("Common gain adjust" and "fast gain adjust") were set to 0dB for this measurement. I didn't need any ad-hoc poles and zeros for this fit (i.e. I can map all the fitted poles and zeros to the schematic), but the fit starts to deviate from the measurement just below 1 MHz.. perhaps I need to add a zero above 1MHz, but I can't see why from the schematic...

Attachment #4: Measured TF from "TEST 1 IN" to "PC OUT" on the FSS box. MEDM gains were once again 0dB. I can't get a good fit to this, mainly because I can't decipher the poles and zeros for this path from the schematic (there are actually deviations from the schematic posted on the 40m DCC page in terms of component values, I will try and correct whatever I notice. I'll work on this...

Attachment #5: Data files + .fil files used to fit the data with LISO

Quote:

 

Data + plots + fits + updated schematics to follow...

Most of the model has come together, I am not too far from matching the modelled OLG to the measured OLG. So I will now start thinking about designing the controller for the MCL part (there are a couple of TFs that have to be measured for this path).

Attachment 1: mc40_v1.pdf
mc40_v1.pdf
Attachment 2: CMboard_OLTF_fit.pdf
CMboard_OLTF_fit.pdf
Attachment 3: FSSFast_OLTF_fit.pdf
FSSFast_OLTF_fit.pdf
Attachment 4: PCdrive_OLTF_measured.pdf
PCdrive_OLTF_measured.pdf
Attachment 5: data.zip
  12816   Fri Feb 10 02:14:10 2017 gautamUpdateIMC29.5 MHz stabilizer box replacement

Lydia finished up installing the new RF amplifier, and will elog the details of the installation.

I wanted to try and measure the IMC OLG to compare against my Simulink model. So I went about performing a few checks. Summary of my findings:

  1. The amplifier seems to be working fine. I checked powers at the input, output to EOM and output to distribution box (that serves the various LOs) first with a 30dB attenuator at the input, and subsequently with the design choice of 5dB attenuator at the input. Everything seemed in order.
  2. I installed a 30 dB attenuator at the MC REFL PD input to the demod board since my (rough) calculations suggested that our modifications would have resulted in the RF beat power between carrier and sideband increasing in power by ~27dB.
  3. I then opened the PSL shutter and tried locking the IMC - with manual tweaking of the various gains, I was able to lock.
  4. But getting to this point took me a while so I couldn't get an OLG measurement in.

TBC tomorrow, I'm leaving the PSL shutter closed and the RF source off for tonight...

  12820   Fri Feb 10 18:21:21 2017 gautamUpdateIMCIMC Demod board

Rana and I spent some time looking at the IMC demod board earlier today. I will post the details shortly, but there was a label on the front panel which said that the nominal LO level to the input should be -8dBm. The new 29.5MHz routing scheme meant that the LO board was actually being driven at 0dBm (that too when the input to the RF distribution box was attenuated by 5dB).

An elog search revealed this thread, where Koji made some changes to the demod board input attenuators. Rana commented that it isn't a good idea to have the LO input be below 0dBm, so after consulting with Koji, we decided that we will

  • Remove the 5dB attenuator to the input of the distribution box such that the LO is driven at ~5dBm
  • Remove the input 10dB attenuator, first ERA-5SM amplifier, and the mini circuits power splitter from the demod board (schematic to follow).

After implementing these changes, and testing the board with a Marconi on the workbench, I found that the measured power levels (measured with an active FET probe) behave as expected, up till the ERA-5SM immediately prior to the LO (U4 and U6 on the schematic). However, the power after this amplifier (i.e. the input to the on-circuit LO, Minicircuits JMS-1H, which we want to be +17dBm), is only +16dBm. The input to these ERA-5SMs, which are only ~2years old, is -2dBm, so with the typical gain of +20dB, I should have 18dBm at their output. Moreover, increasing the input power to the board from the Marconi doesn't linearly increase the output from the ERA-5SM. Just in case, I replaced one of the ERA-5SMs, but observed the same behaviour, even though the amplifier shouldn't be near saturation (the power upstream of the ERA-5SM does scale linearly).

This needs to be investigated further, so I am leaving the demod board pulled out for now...

  12822   Sun Feb 12 01:16:57 2017 gautamUpdateIMCIMC length loop - summary of changes

29.5 MHz RF Modulation Source

  • The +13dBm from the Wenzel oscillator gets amplified to +27dBm by a ZHL-2-S. There is a 5dB attenuator on the input to the amplifier to avoid compression/saturation.
  • The amplified output goes to the EOM (+26dBm measured at the rack, no measurement done at the input to the triple-resonant circuit box yet), while a 10dB coupled part goes to the RF distribution box which splits the input into 16 equal parts. The outputs were measured to spit out +5dBm.
  • 2 of these go to the WFS demod boards - it was verified that this level of drive is okay for the comparator chips on the demod board.
  • A third output goes to the IMC Demod board. The demod board was modified so that the nominal LO input level is now +5dBm (details below).
  • The remaining outputs are all terminated with 50ohms.

IMC Demodulation Board

  • The input attenuator, amplifier and power splitter were removed.
  • Schematic with changes marked and power levels measured, along with a high-res photograph (taken with our fancy new Macro lens + LED light ring) has been uploaded to a page I made to track changes for this part on the DCC (linked to 40m document tree).
  • After making the changes, it was verified that the power levels in the signal chain were appropriate up till the input to the ERA-5SM amplifier directly before the LO. These levels were deemed appropriate, and also scaled in a predictable manner with the input power. As Koji mentions in the previous elog, the dynamically changing input impedance of the mixer makes it difficult to measure the LO level at this point, but I am satisfied that it is within ~1dBm of the nominal +17dBm the mixer wants.
  • The board was further checked for gain imbalance and orthogonality of the I and Q outputs. The graphic below show that there is negligible gain imbalance, but the relative phase between the I and Q channels is ~78 degrees (they should be 90 degrees). Of course this doesn't matter for the IMC locking as we only use the I phase signal, but presumably, we want to understand this effect and compensate for it. 

  • The label on the front panel has been updated to reflect the fact that the nominal LO input is now +5dBm
  • The demodulation phase had changed since the RF signal change was modified - Rana and I investigated this effect on Monday morning, and found that a new ~1.5m long cable was needed to route the signal from the RF distribution box to the LO input of the demod board, which I made. Subsequent modifications on the demod board meant that an extra ~10cm length was needed, so I just tacked on a short length of cable. All of the demodulated signal is now in the I output of the demod board (whereas we had been using the Q output).
  • The graphics below confirms that claim above. Note the cool feature on the digital scopes that the display persistence can be set to "infinity"!
        

I wanted to do a quick check to see if the observed signal levels were in agreement with tests done on the workbench with the Marconi. The mixers used, JMS-1H, have an advertised conversion loss of ~7dB (may be a little higher if we are not driving the LO at +17dBm). The Lissajous ellipse above is consistent with these values. I didn't measure powers with the MC REFL PD plugged into the demod board, but the time series plot above suggest that I should have ~0dBm power in the MC REFL PD signal at 29.5MHz for the strongest flashes (~0.3Vpp IF signal for the strong flashes). 

 

MC Servo Board

  • As mentioned above, we now use the I phase signal for lMC PDH locking.
  • This has resulted in an overall sign change of the servo. I have updated the MEDM screen to reflect that "MINUS" is the correct polarity now..
  • To set the various gains, I measured the OLTF for various configurations using the usual IN1/IN2 prescription on the MC Servo Board (using the Agilent analyzer). 
  • I started at 0dBm "In1 Gain", and the nominal (old) values for "VCO gain", "FSS Common Gain" and "FSS FAST gain"  and found that though I could lock the MC, I couldn't reliably turn on the boosts.
  • After some tweaking, I settled on +10dB "In1 Gain". Here, locking was much more reliable, and I was able to smoothly turn on the Super Boosts. The attached OLTF measurement suggests a UGF of ~118kHz and phase margin of a little more than 30 degrees. There is room for optimization here, since we have had UGFs closer to 200kHz in the recent past. 
  • I didn't get around to measuring the actual PZT/EOM crossover yesterday. But I did measure the OLTF for various values of the FSS gains. At the current value of +20dBm, the PC drive signal is hovering around 1.5V. This bit of optimization needs to be done more systematically. 
  • I've edited mcup and mcdown to reflect the new gains. 

Some general remarks

  • The whole point of this exercise was to increase the modulation depth for the 29.5MHz signal. 
  • By my estimate, assuming 8mrad/V modulation index for the EOM and a gain of 0.6 at 29.5 MHz in the triple resonant box, we should have 100mrad of modulation after installing the amplifier (compared to 4mrad before the change). 
  • The actual RF power at 29.5 MHz at the input/output of the triple resonant box has not yet been measured. 
  • The WFS input error signal levels have to be re-measured (so I've turned off the inputs to the digital WFS filters for now)
Attachment 1: DemodBoardOrthogonality.pdf
DemodBoardOrthogonality.pdf
Attachment 2: IMC_PDH.pdf
IMC_PDH.pdf
Attachment 4: IMC_OLTF.pdf
IMC_OLTF.pdf
Attachment 5: FSS_gain_comparison.pdf
FSS_gain_comparison.pdf
  12824   Mon Feb 13 13:34:44 2017 gautamUpdateIMCIMC length loop - bad SMA cable replaced

I was a little confused why the In1 Gain had to be as high as +10dB - before the changes to the RF chain, we were using +27dB, and we expect the changes made to have increased the modulation depth by a factor of ~25, so I would have expected the new In1 Gain to be more like 0dB.

While walking by the PSL table, I chanced upon the scope monitoring PMC transmission, and I noticed that the RIN was unusually high (see the scope screenshot below). We don't have the projector on the wall anymore, but it doesn't look like this has shown up in the SLOW monitor channel anyways. Disabling the MC autolocker / closing the PSL shutter had no effect. I walked over to the amplifier setup in 1X2, and noticed that the SMA cable connecting the output of the amplifier to the EOM drive was flaky. By touching the cable a little, I noticed that the trace on the scope appeared normal again. Turning off the 29.5MHz modulation source completely returned the trace to normal.

 

So I just made a new cable of similar length (with the double heat shrink prescription). The PMC transmission looks normal on the scope now. I also re-aligned the PMC for good measure. So presumably, we were not driving the EOM with the full +27dBm of available power. Now, the In1 Gain on the MC servo board is set to +2dB, and I changed the nominal FSS FAST gain to +18dB. The IMC OLTF now has a UGF of ~165kHz, though the phase margin is only ~27 degrees.. 

Quote:

MC Servo Board

  • After some tweaking, I settled on +10dB "In1 Gain". Here, locking was much more reliable, and I was able to smoothly turn on the Super Boosts. The attached OLTF measurement suggests a UGF of ~118kHz and phase margin of a little more than 30 degrees. There is room for optimization here, since we have had UGFs closer to 200kHz in the recent past. 
  12828   Tue Feb 14 10:43:06 2017 gautamBureaucracyEquipment loanEquipment to Cryo Lab

PZT Buzzer Box (Thorlabs HV Supply + Manual + 2*PZT Buzzers) ---> Cryo Lab (Brittany + Aaron)

  12833   Wed Feb 15 23:54:13 2017 gautamUpdateIMCIMC saga continues...

Following the discussion at the meeting today, I wanted to finish up the WFS tuning and then hand over the IFO to Johannes for his loss stuff. So I did the following:

  1. First I set the dark offsets on the WFS (with PSL shutter closed). Then I hand aligned the MC to maximize transmission, centered the beam on the WFS, and set the RF offsets with the MC unlocked.
  2. Given that the demod phase for the IMC PDH demodulation board changed by |45 degrees|, I tried changing the digital demod phases in each of the WFS quadrant signals by +/- 45 degrees. Turns out +45 degrees put all the error signal into the I Phase, which is what we use for the WFS loops.
  3. Then I attempted to check the WFS loops. I estimated that we have ~25 times the modulation depth now, so I reduced the WFS1/2 P/Y gains by this factor (but left the MC2 TRANS P/Y gains as is). The loop gain seemed overall too low, so I upped the gain till I saw instability in the loop (error signals ringing up). Then I set the loop gains to 1/3 of this value - it was 0.01 before, and I found the loop behaved well (no oscillations, MC TRANS stabilized) at a gain of 0.002.

At this point, I figured I would leave the WFS in this state and observe its behaviour overnight. But abruptly, the IMC behaviour changed dramatically. I saw first that the IMC had trouble re-acquiring lock. Moreover, the PC Drive seemed saturated at 10.0V, even when there was no error signal to the MC Servo board. Looking at the MEDM screen, I noticed that the "C1-IOO_MC_SUM_MON" channel had picked up a large (~3V) DC offset, even with In1 and In2 disabled. Moreover, this phenomenon seemed completely correlated with opening/closing the PSL shutter. Johannes and I did some debugging to make sure that this wasn't a sticky button/slider issue, by disconnecting all the cables from the front panel of the servo board - but the behaviour persisted, there seemed to be some integration of the above-mentioned channel as soon as I opened the PSL shutter.

  

 

Next, I blocked first the MC REFL PD, and then each of the WFS - turns out, if the light to WFS2 was blocked and the PSL shutter opened, there was no integrating behaviour. But still, locking the MC was impossible. So I suspected that something was wrong with the LO inputs to the WFS Demod Boards. Sure enough, when I disconnected and terminated those outputs of the RF distribution box, I was able to re-lock the MC fine.

I can't explain this bizzare behaviour - why should an internal monitor channel of the MC Servo board integrate anything when the only input to it is the backplane connector (all front panel inputs physically disconnected, In1 and In2 MEDM switches off)? Also, I am not sure how my work on the WFS could have affected any hardware - I did not mess around at the 1X1 rack in the evening, and the light has been incident on the WFS heads for the past few days. The change in modulation depth shouldn't have resulted in the RF power in this chain crossing any sort of damage threshold since the measured power before the changes was at the level of -70dBm, and so should be at most -40dBm now (at the WFS demod board input). The only thing different today was that the digital inputs of the WFS servos were turned on...

So for tonight I am leaving the two outputs of the RF distribution box that serve as the LO for the WFS demod boards terminated, and have also blocked the light to both WFS with beam blocks. The IMC seems to be holding lock steady, PC drive levels look normal...


Unrelated to this work, but I have committed to the svn the updated versions of the mcup and mcdown scripts, to reflect the new gains for the autolocker...

  12834   Thu Feb 16 13:29:38 2017 gautamSummaryGeneralAlternative Calibration Scheme

Summary:

Craig and I have been trying to put together a Simulink diagram of the proposed alternative calibration scheme. Each time I talk the idea over with someone, I convince myself it makes sense, but then I try and explain it to someone else and get more confused. Probably I am not even thinking about this in the right way. So I am putting what I have here for comments/suggestions.

What's the general idea?

Suppose the PSL is locked to the MC cavity, and the AUX laser is locked to the arm cavity (with sufficiently high BW). Then by driving a line in the arm cavity length, and beating the PSL and AUX lasers, we can determine how much we are modulating the arm cavity length in metres by reading out the beat frequency between the two lasers, provided the arm cavity length is precisely known.

So we need:

  1. Both lasers to be stabilized to be able to sense the line we are driving
  2. A high bandwidth PDH loop for locking the AUX laser to the arm cavity such that the AUX laser frequency is able to track the line we are driving
  3. An accurate and precise way to read out the beat frequency (the proposal here is to use an FPGA based readout)
  4. An accurate measurement of the arm length (I think we know the arm lengths to <0.1% so this shouldn't dominate any systematic error).

To be able to sense a 1kHz line being driven at 1e-16 m amplitude, I estimate we need a beat note stability of ~1mHz/rtHz at 1kHz.

Requirements and what we have currently:

  • The PSL is locked to the mode-cleaner, and the arm cavity is locked to the PSL. The former PDH loop is high BW, and so we expect the stabilized PSL to have frequency noise of ~1mHz/rtHz at about 1kHz (to be measured and confirmed)
  • The AUX laser is locked to the arm cavity with a medium-BW (~10kHz UGF) PDH servo. From past out-of-loop ALS beat measurements, I estimate the expected frequency noise of the AUX laser at 1kHz to be ~1Hz/rtHz with the current PDH setup
  • Rana suggested we "borrow" the stability of the PSL by locking the AUX laser and PSL in a high bandwidth PLL - if we want this loop to have ~300kHz BW, then we need to use an EOM as an actuator. The attached Simulink diagram (schematic representation only, though I think I have measurements of many of those transfer functions/gains anyways) shows the topology I had in mind. Perhaps I did not understand this correctly, but if we have such a loop with high gain at 1kHz, and the error signal being the beat between PSL and AUX, won't it squish the modulation we are applying @1kHz?
  • Is it feasible to instead add a parallel path to the end PDH loop with an EOM as an actuator (similar to what we do for the IMC locking)? Ideally, what we want is an end PDH loop which squishes the free-running NPRO noise to ~1mHz/rtHz at 1kHz instead of the 1Hz/rtHz we have currently. This loop would then also have negligible tracking error at 1kHz. Then, we could have a low bandwidth PLL offloading onto the temperature of the crystal to keep the beat between the two lasers hovering around the PSL frequency.

Hardware:

On the hardware side of things, we need:

  • Broadband EOM
  • FSS box to drive the EOM (Rana mentioned there is a spare available in the Cryo lab)

Koji and I briefly looked through the fiber inventory we have yesterday. We have some couplers (one mounted) and short (5m) patch fibers. But I think the fiber infrastructure we have in place currently is adequate - we have the AUX light brought to the PSL table, and there is a spare fiber running the other way if we want to bring the PSL IR to the end as well.

I need to also think about where we can stick the EOM in given physical constraints on the EX table and the beam diameter/aperture of EOM...

Attachment 1: AltCal.pdf
AltCal.pdf
  12838   Fri Feb 17 20:10:18 2017 gautamUpdateIMCWFS servos turned back on

[Koji, gautam]

Turns out the "problem" with WFS2 and the apparent offset accumulation on the IMC Servo board is probably a slow machine problem.

Today, Koji and I looked at the situation a little more closely. This anomalous behaviour of the C1:IOO-MC_SUM channel picking up an offset seems correlated with light being incident on WFS2 head. Placing an ND filter in front of WFS 2 slowed down the rate of accumulation (though it was still present). But we also looked at the in-loop error signal on the IMC board (using the "Out 2" BNC on the front panel), and this didn't seem to show any offset accumulation. Anyways, the ability of the Autolocker doesn't seem to be affected by this change, so I am leaving the WFS servo turned on.

The new demod phases (old +45degrees) and gains (old gains *0.2) have been updated in the SDF table. It remains to see that the WFS loops don't drag the alignment over longer timescales. I will post a more detailed analysis here over the weekend...

Also, we thought it would be nice to have DQ channels for the WFS error signals for analysis of the servo (rather than wait for 30 mins to grab live fine resolution spectra of the error signals with the loop On/Off). So I have added 16 DQ channels [recorded at 2048 Hz] to the c1ioo model (for the I and Q demodulated signal from each quadrant for the 8 quadrants). The "DRATE" for the c1ioo model has increased from ~200 to 410. Comparing to the "DRATE" of c1lsc, which is around 3200, we think this isn't significantly stretching the DAQ abilities of the c1ioo model...

 

  12840   Sat Feb 18 21:50:48 2017 gautamUpdateIMCWFS servos turned back on

Here is a comparison of the error signal spectra after increasing the IMC modulation depth, to the contribution with RF inputs / whitening inputs terminated (which I borrowed from Koji's characterization of the same in Dec 2016, these shouldn't have changed).

Some general observations:

  1. This data was taken with the WFS servos disabled, but with the IMC hand-aligned to a good state (MC_TRANS ~15,000). The error signal spectra are from the new DQ channels (but still sampled at 2048Hz, I had not implemented the change to 512Hz).
  2. The error signals seem to have increased by ~25x yes, which is consistent with how much we expect the modulation depth to have increased
  3. The bump around 1 Hz is now cleaerly visible in all 16 channels, as is the bounce peak at 16Hz (relative to Dec 2016). In general, between 0.1Hz and 5Hz, there is now a fair bit of daylight between the error signals and the electronics noise contribution. 

I will update with the in-loop error signal spectra, which should give us some idea of the loop bandwidth.


I will look into lowering the sampling rate, and how much out-of-band power is aliasing into the 0-256 Hz band and update with my findings.

Quote:

Yikes. Please change the all teh WFS DQ channels sample rates from 2048 down to 512 Hz. I doubt we ever need anything about 180 Hz.

There is sometimes an issue with this: if our digital AA filters are not strong enough, the noise about above 256 Hz can alias into the 0-256 Hz band. We ought to check this quantitatively and make some elog statement about our AA filters. This issue is also seen in DTT when requesting a low frequency spectrum: DTT uses FIR filters which are sometimes not sharp enough to prevent this issue.

 

 

Attachment 1: WFS_error_noise.pdf
WFS_error_noise.pdf