  5942   Fri Nov 18 02:50:10 2011 kiwamuUpdateIOORF generation box : power switch malfunction

[Suresh / Kiwamu]

 The power switch button of the RF generation box is not properly working

For tonight we are leaving it as it is but it needs to be fixed at some point.

 

(the Story)

While I was working around the green broad-band RFPD, I noticed that the RFPD was detecting the 25 MHz modulation signal.
To confirm if it really comes from the modulation source, I switched OFF the RF generation box by pressing the blue LED power button on the rear side of it.
The 25 MHz signal in the RFPD disappeared. So it was indeed the 25 MHz modulation signal.
Then I pressed the LED button again to bring it ON, but the switch didn't stay in the clicked position.
Keeping pressing the button could make it ON but once I released my finger from it it became OFF.
So the mechanical thing in the LED button is not properly working.
I removed the box from the 1X2 rack to take a look at it.
With a help from Suresh we somehow managed to keep it ON after several trials of pressing it.

The temporary solution we decided is to leave it ON so that we can survive tonight.

The box was back in place. The MC is find and 11 MHz and 55 MHz seem okay.

 

Please be aware of it.

 

broken_power_switch.png

This is a picture showing the rear view of the RF generation box. The red arrow is pointing the blue LED switch button.

  5947   Fri Nov 18 15:35:18 2011 kiwamuUpdateIOORF generation box : power switch malfunction

Jenne gave me a spare LED power switch .

I will replace the broken one on Monday.

By the way here is a picture album of the RF generation box which I took last night.

            

Quote from #5942

 The power switch button of the RF generation box is not properly working

  11605   Wed Sep 16 03:44:18 2015 KojiUpdateLSCRF micky mouse

1. POP110 RF amps are powered from the cross connect. But that +15V block has GND connections that are not connected to the ground.
    i.e. The ground potential is given by the signal ground. (Attachment 1)

    This is caused by the misuse of the DIN connector  blocks. The hod side uses an isolated block assuming a fuse is inserted.
    However, the ground sides also have the isolated blocks

2. One of the POP110 RF cable has a suspicious shiled. The rigidity of the cable is low, suggesting the broken shield. (Attachment 2)

Attachment 1: IMG_20150915_231038191.jpg
IMG_20150915_231038191.jpg
Attachment 2: IMG_20150915_234257144.jpg
IMG_20150915_234257144.jpg
  11613   Thu Sep 17 17:27:01 2015 gautamUpdateLSCRF micky mouse - dodgy DIN connector blocks fixed

[Steve, gautam]

We fixed the problematic DIN connectors on 1Y2, by swapping out the 3 DIN connector blocks that were of the wrong type (see attached image for the difference between the types appropriate for "Live" and "Ground").

Before doing anything, Eric turned the Wenzel multiplier off. We have not turned this back on.

Then we turned off the power supply unit at the base of 1Y2, removed the connectors from the rail, swapped out the connectors, reinstalled them on the rail, and turned the power supply back on. After swapping these out, we verified with a multimeter that between each pair of "Live" and "Ground" blocks, there was ~15V. We could now use the third unused pair of blocks to power the delay line phase shifter box, though for the moment, it remains powered by the bench power supply. 

Quote:

1. POP110 RF amps are powered from the cross connect. But that +15V block has GND connections that are not connected to the ground.
    i.e. The ground potential is given by the signal ground. (Attachment 1)

    This is caused by the misuse of the DIN connector  blocks. The hod side uses an isolated block assuming a fuse is inserted.
    However, the ground sides also have the isolated blocks

2. One of the POP110 RF cable has a suspicious shiled. The rigidity of the cable is low, suggesting the broken shield. (Attachment 2)

 

Attachment 1: DIN_rail_terminal.jpg
DIN_rail_terminal.jpg
  11614   Thu Sep 17 19:42:43 2015 KojiUpdateLSCRF micky mouse - dodgy DIN connector blocks fixed

1. The delay-line box is now hooked up to the cross connect +15V supply.

2. The broken RF cable was fixed.

It is actually the POP22 cable.
Therefore, we might see significant change of the signal size for POP22.
Be aware.

RG405 + SMA connector rule

- Don't bend the cable at the connector.

- Always use a cap on the connector. It is a part of the impedance matching.

- Use transparent shrink tube for strain relieving and isolation. This allow us to check the condition of the shield without removing the cover.

Attachment 1: IMG_20150917_190635033.jpg
IMG_20150917_190635033.jpg
Attachment 2: IMG_20150917_192551919.jpg
IMG_20150917_192551919.jpg
  11616   Fri Sep 18 08:03:53 2015 ranaUpdateLSCRF micky mouse - dodgy DIN connector blocks fixed

Steve and I turned on the box this morning so that the IMC would lock again.

For future reference, remember that one should turn off the Marconi output before turning off the RF distribution box. Don't drive the input of unpowered RF amps.

 

  11617   Fri Sep 18 08:04:09 2015 ranaUpdateLSCRF micky mouse - dodgy DIN connector blocks fixed

Steve and I turned on the box this morning so that the IMC would lock again.

For future reference, remember that one should turn off the Marconi output before turning off the RF distribution box. Don't drive the input of unpowered RF amps.

 

  5420   Thu Sep 15 17:12:29 2011 MirkoUpdateLSCRF modulation depth measurement

[Mirko, Kiwamu]

Put up a temp. setup on the laser table to measure the RF modulation depths using the optical spectrum analyzer. First with a pickoff beam with about 2mW => SNR of 8 of 1 peak per FSR.

Then with a beam with about 100mW. Much better SNR on the single peak but still no sidebands visible. Modematching not too good in either case. Shouldn't matter.

  5462   Mon Sep 19 15:44:32 2011 MirkoUpdateLSCRF modulation depth measurement

Earlier measurements of the modulation index were less than optimal because we had too low transmission through the cavity. Contrary to what was believed you actually need to modematch onto the cavity.

Earlier transmitted power was about 8.5uW.

With a 250mm lens we archived 41uW.

Impinging power on the cavity is 1.7mW.

PD TF approx 0.1V / uW.

Carrier power: 4.1V => 41uW

41uW/1.7mW = 2.4 % transmission. Manufacturer clain for peak transmission: 20-30%.

11MHz SB: 28.8mV => m=0.17

55MHz SB:36mV => m=0.19


As you can see on the pic the SNR of the SBs is not too good.

P9190138.JPG

  5519   Thu Sep 22 15:53:37 2011 MirkoUpdateLSCRF modulation depth measurement again

Toyed around with the modematching some more today.

The outermost glass elements of the OSA are about 28cm apart.

With the OSA beeing a confocal cavity that should mean that the ROC of every mirror is 28cm on the cavity side. If the input surface is flat we need a 28cm focusing lens for good MM. If it's not we shouldn't need any MM.

Tried a f=250mm lens on different positions first. Got at best about 570mV (PD gain=10) in transmission of the OSA.

Then tried a f=1000mm lens. Best transmission 1.22V (7.2% transmission). SB were (PD gain =100) 11MHz: 87.2mV (m=0.17) , 55MHz: 59.2mV (m=0.14)

Lost the position while toying around. Left it then at 1.0V transmisison at 15:15 local time. Let's see how much it drifts. SBs for this were 11MHz: 52.8mV (m=0.17), 55MHz: 73.8mV (m=0.14)

[Ed by KA: If the carrier transmission was really 1.22V and 1.0V the modulation depths calculated are inconsistent with the measurement.]

Spacing between carrier 11MHz and 55MHz SBs seems consistent, and leads to a FSR measurement of 1.5GHz, also fine.

 

Update: After 90mins no change in carrier transmitted power. Next morning: Carrier transmission down by 10%.

DSC_3478.JPG

 DSC_3481.JPG

DSC_3480.JPG

  285   Wed Jan 30 11:49:30 2008 AlbertoSummaryElectronicsRF monitor's filters final schematics and transfer functions
These are the final schematics for the 6th order Chebyshev filters of the RF monitor board. I'm also attaching the TF as I measured. The tuning is probably not optimal, less insertion noise could be achieved.
Attachment 1: 33Melog30Jan08.png
33Melog30Jan08.png
Attachment 2: 133Melog30Jan08.png
133Melog30Jan08.png
Attachment 3: 166Melog30Jan08.png
166Melog30Jan08.png
Attachment 4: 199Melog30Jan08.png
199Melog30Jan08.png
Attachment 5: 33elog30Jan08.png
33elog30Jan08.png
Attachment 6: 133elog30Jan08.png
133elog30Jan08.png
Attachment 7: 166elog30Jan08.png
166elog30Jan08.png
Attachment 8: 199elog30Jan08.png
199elog30Jan08.png
  15324   Mon May 11 00:12:34 2020 gautamUpdateLSCRF only PRFPMI

Finally - Attachment #1. This plot uses 16 Hz EPICS data. All y-axes are uncalibrated for now, but TRX/TRY are normalized such that the POX/POY lock yields a transmission of 1. CARM UGF is only ~3 kHz, no boosts were turned on yet. 

Attachment #2 and Attachment #3 are phone photos of the camera images of the various ports. After some alignment work, the transmitted arm powers were ~200, i.e. PRG ~10. fwiw, this is the darkest i've ever seen the 40m dark port. c.f. 2016. Of course, the exposure time / ND filter / light levels could all have changed.

This work was possible during the daytime (~6pm PDT), but probably only because it was Sunday. The other rate limiting factor here is the franky terrible IMC duty cycle. TBH, I didn't honestly expect to get so far and ran out of time, but I think the next steps are:

  1. Turn on some sensing lines and calibrate CARM/DARM.
  2. Transition vertex control to 1f signals.
  3. Whiten DARM.
  4. Turn on some ASC for better power stabilization.
  5. Scan the CARM offset and check that we are truly on resonance
  6. Noise budget.

As usual, I would like to request that we don't change the IFO as far as possible until the BHD vent, i found it pretty difficult to get here.


Attachment #4 now shows the measured DARM OLTF when DARM is entirely on AS55_Q control. UGF is ~120 Hz and the phase margin is ~30 deg, seems okay for a first attempt. I'll now need to infer the OLTF over a wider range of frequencies by lining this measurement up with some model, so that I can undo the loop in plotting the DARM ASD.

Attachment 1: PRFPMIlock.pdf
PRFPMIlock.pdf
Attachment 2: IMG_8549.JPG
IMG_8549.JPG
Attachment 3: IMG_8548.JPG
IMG_8548.JPG
Attachment 4: DARM_OLTF.pdf
DARM_OLTF.pdf
  926   Thu Sep 4 17:03:25 2008 YoichiUpdatePSLRF oscillator noise comparison
I measured current spectra of the RF signal going to the FSS EOM.
The attachment compares the spectra between a Stanford signal generator and a Marconi.
I borrowed the Marconi from the abs. length measurement experiment temporarily.
The measurement was done using the signal going to the EOM. That means the spectra include
noise contributions from the RF amp., splitter and cables.

21.5MHz peak was not included because that would overload the ADC and I would have to use a large attenuation.
This means the measurement would be totally limited by ADC noise everywhere except for 21.5MHz.

I noticed that with the Marconi, the FSS is a little bit happier, i.e. the PC path is less loaded
(0.9Vrms with Stanford vs. 0.7Vrms with Marconi). But the difference is small.
Probably the contribution from the 77kHz harmonics in the laser light is more significant (see entry #929).
Also the peaks in the Stanford spectrum are not harmonics of 77kHz, which we see in the FSS error signal.

I returned the Marconi after the measurement to let Alberto work on the abs. length measurement.
Attachment 1: RFSpectra.png
RFSpectra.png
  4675   Tue May 10 01:39:41 2011 SureshUpdateRF SystemRF source troubles

Today after Steve and I finished the RF cabling work for the day, Kiwamu noticed that there were no RF signals to be seen.  The problem was traced to disconnected 11 and 55 MHz Demod lines from the RF source.  But reconnecting them did not restore the signals.  It turned out that one of the Heliax cables had a loose N-type connector at its end and it finally came off while we were tightening it into place.

We replaced the damaged heliax with another (we have two spare running from 1X2 IOO rack to the 1Y2 LSC rack.  The new cable is used to be the LO 33.  It seems to have a 1.5dB loss.  Have to check this again tomorrow.

In the mean time I noticed that the power output of the 55MHz Demod port of the source was less than about -12dB. So I opened the source to take a look and found that all the voltage stabilisers were supplying 15V.  Even those which were supposed to be supplying 24V.  This was traced to a mistake in wiring the external power supply.  The wires had been labeled wrongly and as a result the 18V input line was connected to 28V source and vice versa.

After fixing this problem I reassembled the source checked the power output on all the ports and found everything was functioning as expected.  However after installation once again the unit failed.  The blue light on the power supply was not lighting up when switched on.  Suspecting a power supply problem I opened the unit again and found that a weak solder joint on one of the RF amplifiers had come loose and had overloaded one of the 24V stabilisers.   We, found a spare and replaced it.  The unit has been reassembled and is functioning fine.  The output power levels are

11MHz Demod -- 6dBm

55MHz Demod -- 5.5 dBm

11MHz EOM --   24dBm

55MHz EOM -- 28dBm

The Marconi is serving as the 11MHz source.  The Wenzel 11MHz source is giving 13.3 dBm and is okay.  But it needs to be checked for its performance as it may have been exposed to higher than rated power supply levels.

The 29.5MHz source is giving 7dBm.  It is supposed to be giving 13dBm. 

The Laboratory DC power supplies currently used for both the RF source and Distribution boxes need to be replaced with rack mounted Sorensen power supplies available in the lab.

 

  4539   Mon Apr 18 14:11:44 2011 kiwamuUpdateLSCRF status

 We will make them all green !!

 RF_Work_Status.png

Again, all the files are available in the svn.

https://nodus.ligo.caltech.edu:30889/svn/trunk/suresh/40m_RF_upgrade/

  4591   Fri Apr 29 18:24:05 2011 SureshUpdateRF SystemRF system: 1X2 Rack cabling

[Joe, Jamie, Suresh]

We have installed the IDE to SCSI adaptor module into the 1X2 rack and have connected the AA filter outputs to it.

P4290070.JPG

 

We have removed the following cables running between the 1X2 and 1X3 racks.

The long twisted pair ribbon cable which previously carried the ADC signals.

1X2-ASC 6, 1X2-ASC 47, 1X2-ASC 9, 1X2-ASC 8, 1X2-ASC 10, 1X2-ASC 7,

CAB-1X2-LSC 42, CAB 1X2-LSC 56,  CAB 1X2-LSC 41, CAB 1X2-LSC 43

1X3-2 ASC 47

We have also removed the following by mistake.  We will put them back them on Monday

1X2-LSC 21, 1X2-LSC-20.

We have also removed the ASC QPD cables and moved the QPD cards which were present in the middle Eurocate (#2) to the unused Eurocrate at the bottom position (#3).

The binary input cables at the back of the cards require to be supported so that their weight does not pull them out of the sockets at the back of the crates.

Some of the slots where we plan to plug in Demod boards (the 165 MHz boards)  are not currently connected to any binary output on the C1:LSC computer.  We need these binary controls for the fitlter modules on the cards.

When we eventually begin to use the 15PDs as planned, then we will occupy 30 ADC channels (I & Q outputs).  Currently we have just one ADC card installed on the C1:LSC providing 32 ADC channels.  Joe found another 16bit 32 channel ADC card in his stash but we need to get a timing+adaptor board for it. In general we are going to need the third Eurocrate.

A platform for the power supply of the RF Distribution box needs to be built and the power supply needs to be moved into the 1X2 rack rather than sit on top of 1X2 rack.

 

 

 

  4547   Wed Apr 20 21:53:01 2011 SureshConfigurationRF SystemRF system: Stray heliax cable

We found a stray unused heliax cable running from the LSC rack 1Y2 to a point between the cabinets 1X3 and 1X4. This cable will need to be redirected to the AS table in the new scheme.   It is labled C1LSC-PD5  The current situation has been updated as seen in the layout below

rogue_cable_1.png

Attachment 1: rogue_cable_1.png
rogue_cable_1.png
  4330   Sat Feb 19 05:25:20 2011 SureshUpdateElectronicsRF: Distribution box

Most of the RF cables required for the box are done.   There are two remaining and we will attend to these tonight. 

We expect to have finished the mechanical assembly by Sunday and start a quality test on Monday.

 

 

  361   Wed Mar 5 17:35:24 2008 ranaUpdateIOORFAM during MC lock
I used an ezcaservo command to adjust the offsets for Alberto's StochMon channels. They are all
at +2 V with no light
on the RFAM PD (MC unlocked).

Then I looked at 5 minutes of second trend around when the MC locks. Since Alberto has chosen
to use +2V to indicate zero RF and a negative gain, there is a large RF signal when the StochMon
channels approach zero
.

From the plot one can see that the RFAM for the 133 & 199 MHz channels is much worse than for the 33 and 166.
Its also clear that the turn on of the WFS (when the RFAMPD's DC light level goes up) makes the single demod
signals get better but the double demod get worse.
Attachment 1: rfam.pdf
rfam.pdf
  6044   Tue Nov 29 22:10:18 2011 kiwamuUpdateRF SystemRFAM fluctuation reduced

Quote from #6035

I left the EOM stabilization running overnight, so we can finally see how the EOM temperature stabilization does over long periods of time.

The controller was turned on at ~8:40 UTC, and you can see that the Stochmon signals quiet down a lot right at that time. 

Indeed the fluctuation of the RFAM became quieter with the temperature control ON.

However the absolute value of the RFAMs stayed at relatively high value.

I guess we should be able to set the right temperature setpoint such that the absolute value of the RFAM is smaller.

Here is the calibrated RFAM data (for 5 hours around the time when Zach activated the temperature control last night):

RFAM_withEOMheater_edit.png

  6047   Tue Nov 29 23:03:34 2011 ZachUpdateRF SystemRFAM fluctuation reduced

I was hesitant to claim that this is definitely true without the control data we were taking after the heater was turned off today. This is because before I replaced the malfunctioning op amp last night, the heater was actually ON and injecting temperature noise into the system that would not be there with it off. I think the best idea is to compare the data from today (heater on vs. heater off, but with functioning circuit).

Quote:

 

Indeed the fluctuation of the RFAM became quieter with the temperature control ON.

  6050   Wed Nov 30 03:01:55 2011 kiwamuUpdateRF SystemRFAM fluctuation reduced

Okay I have turned ON the temperature control at 2:40 AM and will leave it ON for a while.

Quote from #6047

I was hesitant to claim that this is definitely true without the control data we were taking after the heater was turned off today. This is because before I replaced the malfunctioning op amp last night, the heater was actually ON and injecting temperature noise into the system that would not be there with it off. I think the best idea is to compare the data from today (heater on vs. heater off, but with functioning circuit).

 

  5760   Fri Oct 28 20:39:19 2011 MIrkoUpdateLSCRFAM monitor in place. ( Uncalibrated ) EPICS troubles

{Suresh, Jamie, Mirko]

We adapted the Stochmon box to include LP filters at 1.8Hz behind the RMS parts.
Then measured the RMS signals for different RF signal levels at 11.0.65, 29.5, 55.325MHz provided by a RF freq. generator.
As you can see in the data below the suppression of the BP filters of neighboring frequencies is only 35-35dB in power (see also manufacturer specs).

We therefor want to substract crosstalk, by calculating it out. We decided to use C-code in CDS. No computer crashing this time :)

We however ran into the problem that the RMS signal channels are acquired by the slow (EPICS) maschine c1iool0. Channels are (C1:IOO-RFAMPD_33MHZ , -"-133MHZ, -"-166MHZ) and we could not access those in the CDS c1ioo model. Using the EpicsIn block we got an CA.Exception stating that the variable was hosted on multiple servers. We then tried to use the EzcaRead to access the variables. Got an compile error, about the compiler not beeing able to connect all parts. It seems that the EzcaRead left behind a "ghost" part in the model (something with M1:SYS-FOO_BAR which is the default naming of the EzcaRead block) even after we deleted that block. We toyed around with the /opt/rtcds/caltech/c1/chans/daq/C1EDU_IOO.ini  and  /cvs/cds/caltech/target/c1iool0/ioo.db  files. We tried to uncomment the "old" (33,133,166) channels there to get rid of the conflict, but that didn't work.

We want to write the outputs to C1:IOO-RFAMPD_11MHZ , -"-29MHZ, -"-55MHZ EPICS channels.

We had to get the model back from the svn to get it running again.

Attachment 1: MC_DC11MHz.m
Pwr=[-60,-55,-50,-45,-40,-35,-30,-25,-20,-10,-5,0,5,10]';
Voltage11=[2.12,2.10,2.03,1.93,1.83,1.71,1.59,1.47,1.35,1.10,0.97,0.85,0.71,0.61]';

Voltage11=spline(Pwr,Voltage11,linspace(-60,10,15));
PwrSmooth=linspace(-60,10,15);

Voltage29=[2.14,2.14,2.14,2.14,2.14,2.14,2.13,2.12,2.09,1.94,1.84,1.73,1.61,1.49];
Voltage29=spline(Pwr,Voltage29,linspace(-60,10,15));

Voltage55=[2.16,2.16,2.16,2.16,2.16,2.16,2.16,2.15,2.14,2.13,2.10,2.04,1.94,1.83];
... 5 more lines ...
Attachment 2: MC_DC29MHz.m
Pwr=[-60,-55,-50,-45,-40,-35,-30,-25,-20,-15,-10,-5,0,5,10]';
Voltage11=[2.16,2.16,2.16,2.16,2.15,2.16,2.15,2.13,2.10,2.03,1.93,1.81,1.70,1.58,1.46]';

Voltage29=[2.12,2.10,2.03,1.93,1.83,1.70,1.59,1.47,1.34,1.22,1.09,0.97,0.84,0.71,0.61]';

Voltage55=[2.16,2.15,2.16,2.16,2.15,2.15,2.14,2.10,2.0,1.97,1.97,1.77,1.65,1.50,1.37]';

%%  Example 55MHz inj.

Voltage11=[2.00];
... 21 more lines ...
Attachment 3: MC_DC55MHz.m
Pwr=[-60,-55,-50,-45,-40,-35,-30,-25,-20,-15,-10,-5,0,5,10]';
Voltage11=[2.16,2.16,2.16,2.16,2.16,2.15,2.15,2.15,2.15,2.15,2.12,2.09,2.00,1.89,1.78]';

Voltage29=[2.14,2.14,2.14,2.13,2.14,2.13,2.11,2.06,1.98,1.88,1.76,1.64,1.52,1.40,1.27]';

Voltage55=[2.14,2.11,2.05,1.96,1.85,1.73,1.61,1.48,1.36,1.23,1.10,0.98,0.84,0.71,0.61]';

plot(Pwr,Voltage55)

%%
... 17 more lines ...
Attachment 4: RFAMPD.c
double x;
double y;
double z;

double temp1;
double temp2;

double Corrx;
double Corry;
double Corrz;
... 49 more lines ...
  5764   Sun Oct 30 14:08:35 2011 ranaUpdateElectronicsRFAM monitor in place. ( Uncalibrated ) EPICS troubles

Quote:

{Suresh, Jamie, Mirko]

We adapted the Stochmon box to include LP filters at 1.8Hz behind the RMS parts.
Then measured the RMS signals for different RF signal levels at 11.0.65, 29.5, 55.325MHz provided by a RF freq. generator.
As you can see in the data below the suppression of the BP filters of neighboring frequencies is only 35-35dB in power (see also manufacturer specs).

We therefor want to substract crosstalk, by calculating it out. We decided to use C-code in CDS. No computer crashing this time :)

 This is neat idea, but it seems like it would be easier to just add another set of rf BP filters inside of the StochMon box. Luckily, Steve was thinking ahead and ordered extra filters.

  5745   Thu Oct 27 03:32:45 2011 KojiSummaryIOORFAM monitor progress

[Suresh, Mirko, Koji]

A cable from the stochmon box to the cross connect for the EPICS ADCs is installed.

The power supply and the signal outputs are concentrated in a single DSub 9pin connector
that is newly attached on the box.

The connection from the stochmon side of the cable and the EPICS value was confirmed.
The calibration of them looks fine.

To do:
- Once the stochmon box is completed we can immediately test it.
- The EPICS channel names are still as they were. We need to update the database file of c1iool0, the chans file for the slow channel.


The pinout is as following

-------------
| 1 2 3 4 5 |   Female / Inside View
\  6 7 8 9  /
 \---------/

1 - 11MHz Signal
2 - 30MHz Signal
3 - 55MHz Signal
4 - NC
5 - +5V supply
6 - 11MHz Return
7 - 30MHz Return
8 - 55MHz Return
9 - Supply ground

  5964   Sun Nov 20 15:11:09 2011 kiwamuUpdateIOORFAM monitoring test

DO NOT CHANGE THE IFO ALIGNMENT UNTIL TOMORROW MORNING OR FURTHER NOTICE.

Plus, MC has to be kept locked with the WFS.

 

An RFAM measurement is ongoing

 

 Since the Stochmon turned out to be tricky to calibrate the outputs, Koji and I decided to monitor the RFAMs using REFL11 and REFL55 RFPDs while the beam is single-bounced from PRM.
This is, of course, not a permanent RFAM monitor, but at least it gives us a long-term continuous RFAM information for the first time.
Before the measurement I ran the offset zeroing scripts, therefore any offsets from electronics must be tiny in the acquired REFL signals.
The measurement has begun from approximately 3:00 pm.
 
 Also I found C1LSC.ini file again became default (no channels had been acquired).
So I replaced it with an archived ini file and then restarted fb.
  5966   Mon Nov 21 12:48:00 2011 JenneUpdateIOORFAM monitoring test

I don't think I touched/adjusted/whatever anything, but I did open the PSL table ~5-10min ago to measure the size of the Kiwamu-Box, so if the RFAM stuff looks funny for a few minutes, it was probably me.  Just FYI.

  5967   Mon Nov 21 14:15:25 2011 JenneUpdateIOORFAM monitoring test

Quote:

DO NOT CHANGE THE IFO ALIGNMENT UNTIL TOMORROW MORNING OR FURTHER NOTICE.

 [Mirko,  Jenne]

We're playing with the MC OAF, so we're actuating on MC2.  Again, FYI.

  5968   Mon Nov 21 14:35:28 2011 kiwamuUpdateIOORFAM monitoring test

REFL_RFAM.png

 This is a trend for a day long showing the REFL11/55 demod signals, REFLDC (corresponding to the MC transmitted power) and the PSL booth temperatire.

There are sudden jumps in the REFL55_I and REFL11_Q signals around 5:00 AM this morning, also at the same time the temperature suddenly went up.

But the quality of the signal turned out to be not so good because the fluctuation is still within 1 bit of the ADCs,

we have to try it again with a bigger gain in the analog whitening circuit.

Quote:

An RFAM measurement is ongoing

  5972   Mon Nov 21 17:48:36 2011 KojiUpdateIOORFAM monitoring test

Do we care about the AC? I thought what we care is the DC.

  397   Sun Mar 23 10:42:54 2008 ValeraSummaryElectronicsRFAM of the RF stabilization box is measured
I reconstructed Tobin's setup to measure the RFAM after the RF stabilization box in the 166 MHz modulation path.
The setup consisted of the splitter and the mixer followed by the RF low pass filter and the SR560 (gain x100).
The RF level into splitter was 20 dBm. The Mini-Circuits ZLW-3H (17 dBm LO) mixer was used. The LO was taken
straight out of the splitter and the RF path was attenuated by 11 dBm, The DC out of the mixer was 700 mV.
The noise floor was measured with the RF input of the mixer terminated on 50 Ohm. The 45 MHz measurement
in broad band setting looks better than the noise floor at high frequencies. I am not sure what was wrong with
one or both of those measurements. The 9 MHz measurements are above the noise floor.

The RFAM meets the AdvLIGO requirements in the detection band (f > 10 Hz).

The attached zipped files are:
SRS003 9 MHz DC-200 Hz
SRS004 9 MHz DC-26 kHz
SRS006 45 MHz DC-200 Hz
SRS005 45 MHz DC-26 kHz
SRS007 Noise floor DC-200 Hz
SRS008 Noise floor DC-26 kHz
Attachment 1: RFAM.zip
Attachment 2: amplitudenoise.pdf
amplitudenoise.pdf
  5686   Tue Oct 18 15:20:03 2011 kiwamuSummaryIOORFAM plan

[Suresh / Koji / Rana / Kiwamu]

Last night we had a discussion about what we do for the RFAM issue. Here is the plan.

 

(PLAN)

  1. Build and install an RFAM monitor (a.k.a StochMon ) with a combination of a power splitter, band-pass-filters and Wenzel RMS detectors.

       => Some ordering has started (#5682). The Wenzel RMS detectors are already in hands.

  2. Install a temperature sensor on the EOM. And if possible install it with a new EOM resonant box.

      => make a wheatstone bridge circuit, whose voltage is modulated with a local oscillator at 100 Hz or so.

  3. Install a broadband RFPD to monitor the RFAMs and connect it to the StochMon network.

      => Koji's broadband PD or a commercial RFPD (e.g. Newfocus 1811 or similar)

  4. Measure the response of the amount of the RFAM versus the temperature of the EO crystal.

      => to see whether if stabilizing the temperature stabilizes the RFAM or not.

  5.  Measure the long-term behavior of the RFAM.

      => to estimate the worst amount of the RFAM and the time scale of its variation

  6. Decide which physical quantity we will stabilize, the temperature or the amount of the RFAM.

  7. Implement a digital servo to stabilize the RFAMs by feeding signals back to a heater

     => we need to install a heater on the EOM.

  8. In parallel to those actions, figure out how much offsets each LSC error signal will have due to the current amount of the RFAMs.

    => Optickle simulations.

  9. Set some criteria on the allowed amount of the RFAMs

    => With some given offsets in the LSC error signal, we investigate what kind of (bad) effects we will have.

  5991   Wed Nov 23 18:28:09 2011 KojiUpdateIOORFAMPD channels / EOM monitor channels added to DAQ

The following channels have been registered in c1iool0 database, and are now recorded by FB

C1:IOO-RFAMPD_11MHZ
C1:IOO-RFAMPD_29_5MHZ
C1:IOO-RFAMPD_55MHZ
C1:IOO-RFAMPD_DCMON
C1:IOO-EOM_TEMPMON
C1:IOO-EOM_HEATER_DRIVEMON


PROCEDURE

1) The EPICS database file has been edited to rename/add some channels

/cvs/cds/caltech/target/c1iool0/ioo.db

REMOVED
#grecord(ao,"C1:IOO-RFAMPD_VC")
#grecord(ai,"C1:IOO-RFAMPD_TEMP")
#grecord(ai,"C1:IOO-RFAMPD_DCMON")
#grecord(bo,"C1:IOO-RFAMPD_BIAS_ENABLE")
#grecord(bi,"C1:IOO-RFAMPD_BIAS_STATUS")
#grecord(calc, "C1:IOO-RFAMPD_33MHZ_CAL")
#grecord(calc, "C1:IOO-RFAMPD_133MHZ_CAL")
#grecord(calc, "C1:IOO-RFAMPD_166MHZ_CAL")
#grecord(calc, "C1:IOO-RFAMPD_199MHZ_CAL")

ADDED/EDITED
grecord(ai,"C1:IOO-RFAMPD_11MHZ")
        field(DTYP,"VMIVME-3113")                                              
        field(INP,"#C1 S25 @")
...

grecord(ai,"C1:IOO-RFAMPD_29_5MHZ")
        field(DTYP,"VMIVME-3113")                                              
        field(INP,"#C1 S26 @")

...
grecord(ai,"C1:IOO-RFAMPD_55MHZ")
        field(DTYP,"VMIVME-3113")                                              
        field(INP,"#C1 S27 @")

...
grecord(ai,"C1:IOO-RFAMPD_DCMON")
        field(DTYP,"VMIVME-3113")                                              
        field(INP,"#C1 S28 @")

...
grecord(ai,"C1:IOO-EOM_TEMPMON")
                                                
        field(DTYP,"VMIVME-3113")                                               
        field(INP,"#C1 S29 @")

...
grecord(ai,"C1:IOO-EOM_HEATER_DRIVEMON")

        field(DTYP,"VMIVME-3113")                                              
        field(INP,"#C1 S30 @")

2) The channels have been added to the frame builder database

/cvs/cds/rtcds/caltech/c1/chans/daq/C0EDCU.ini

[C1:IOO-RFAMPD_11MHZ]
[C1:IOO-RFAMPD_29_5MHZ]
[C1:IOO-RFAMPD_55MHZ]
[C1:IOO-RFAMPD_DCMON]
[C1:IOO-EOM_TEMPMON]
[C1:IOO-EOM_HEATER_DRIVEMON]

Note that this C0EDCU.ini is the file that has been registered in

/cvs/cds/rtcds/caltech/c1/target/fb/master

3) burt restore request files were updated

RFAM related settings were removed as they don't exist anymore.

/cvs/cds/caltech/target/c1iool0/autoBurt.req
/cvs/cds/caltech/target/c1iool0/
saverestore.req

4) c1iool0 were rebooted. Framebuilder restarted. c1iool0 were burtrestored.

  5995   Thu Nov 24 05:10:00 2011 KojiUpdateIOORFAMPD channels / EOM monitor channels added to DAQ

EOM TEMPMON and HEATER DRIVEMON have been hooked up to the following channels.

C1:IOO-EOM_TEMPMON
C1:IOO-EOM_HEATER_DRIVEMON 

What a fragile circuit...

I found some of the resistors popped up from the board because of the tension by the Pomona grabbers.
I tried to fix it based on the schematic (photo) and the board photo.

  5997   Thu Nov 24 10:27:07 2011 JenneUpdateIOORFAMPD channels / EOM monitor channels added to DAQ

Here is a drawing of where the monitors are coming from:

EOM_temp_sense_heater_drive_schematic_withMONs.png

 Since we can't put current into the ADC, the heater drivemon is measuring the input of the OP27, which is related to the amount of current sent to the heater.

Quote:

EOM TEMPMON and HEATER DRIVEMON have been hooked up to the the following channels.

C1:IOO-EOM_TEMPMON
C1:IOO-EOM_HEATER_DRIVEMON

 

  5998   Thu Nov 24 12:45:12 2011 ZachUpdateIOORFAMPD channels / EOM monitor channels added to DAQ

Jenne: The point you indicate for the heater monitor is a virtual ground--it will be driven to zero by the circuit if it's functioning properly; the readout should be done at the input pin (2, I think) to the BUF634.

Koji: This is odd, as I made a point of not attaching any clips directly to resistors for exactly this reason. I was also careful to trim resistor/capacitor leads so that they were not towering over the breadboard and prone to bending (with the exception of the gain-setting resistor of the AD620, which was changed at the last minute). At the end of the day, it is a breadboard circuit with Pomona "readout", so it's not going to be truly resilient until I put it on a protoboard. Another thing: I think the small Pomona clips are absolutely terrible, since they slip off with piconewtons of tension; I could not find any more regular clips, so I used them against my better judgment.

  5999   Thu Nov 24 13:54:31 2011 KojiUpdateIOORFAMPD channels / EOM monitor channels added to DAQ

Those clips for the readouts were the ones who popped out.
When I have restored the connections, I checked the schematic and the heater drive mon is clipped on the output side of the OP27.

Quote:

Jenne: The point you indicate for the heater monitor is a virtual ground--it will be driven to zero by the circuit if it's functioning properly; the readout should be done at the input pin (2, I think) to the BUF634.

Koji: This is odd, as I made a point of not attaching any clips directly to resistors for exactly this reason. I was also careful to trim resistor/capacitor leads so that they were not towering over the breadboard and prone to bending (with the exception of the gain-setting resistor of the AD620, which was changed at the last minute). At the end of the day, it is a breadboard circuit with Pomona "readout", so it's not going to be truly resilient until I put it on a protoboard. Another thing: I think the small Pomona clips are absolutely terrible, since they slip off with piconewtons of tension; I could not find any more regular clips, so I used them against my better judgment. 

 

  6633   Wed May 9 11:31:50 2012 DenUpdateCDSRFM

I added PCIE memory cache flushing to c1rfm model by changing 0 to 1 in /opt/rtcds/rtscore/release/src/fe/commData2.c on line 159, recompiled and restarted c1rfm.

Jamie, do not be mad at me, Alex told me do that!

However, this did not help, C1RFM did not start. I decided to restart all models on C1SUS machine in hope that C1RFM uses some other models and can't connect to them but this suspended C1SUS machine. After reboot encounted the same C1SUS -> FB communication error and fixed it in the same was as in the previous case of C1SUS reboot. This happens already the second time (out of 2) after C1SUS machine reboot.

I changed /opt/rtcds/rtscore/release/src/fe/commData2.c back, recompiled and restarted c1rfm. Now everything is back. C1RFM -> C1OAF is still bad.

  6635   Wed May 9 15:02:50 2012 DenUpdateCDSRFM

Quote:

However, this did not help, C1RFM did not start. I decided to restart all models on C1SUS machine in hope that C1RFM uses some other models and can't connect to them but this suspended C1SUS machine.

 This happened because of the code bug -

// If PCIE comms show errors, may want to add this cache flushing
#if 1
if(ipcInfo[ii].netType == IPCIE)
          clflush_cache_range (&(ipcInfo[ii].pIpcData->dBlock[sendBlock][ipcIndex].data), 16); // & was missing - Alex fixed this
#endif
 

After this bug was fixed and the code was recompiled, C1:OAF_MCL_IN is OK, no errors occur during the transmission C1:OAF-MCL_ERR=0.

So the problem was in the PCIE card that could not send such amount of data and the last channel (MCL is the last) was corrupted. Now, when Alex added cache flushing, the problem is fixed.

We should spend some more attention to such problems. This time 2046 out of 2048 points were lost per second. But what if 10-20 points are lost, we would not notice that in the dataviewer, but this will cause problems.

  16097   Thu Apr 29 15:11:33 2021 gautamUpdateCDSRFM

The problem here was that the RFM errors cropped up again - seems like it started ~4am today morning judging by TRX trends. Of course without the triggering signal the arm cavity couldn't lock. I rebooted everything (since just restarting the rfm senders/receivers did not do the trick), now arm locking works fine again. It's a bit disappointing that the Rogue Master setting did not eliminate this problem completely, but oh well...

It's kind of cool that in this trend view of the TRX signal, you can see the drift of the ETMX suspension. The days are getting hot again and the temp at EX can fluctuate by >12C between day and night (so the "air-conditioning" doesn't condition that much I guess 😂 ), and I think that's what drives the drift (idk what the transfer function to the inside of the vacuum chamber is but such a large swing isn't great in any case). Not plotted here but i hypothesize TRY levels will be more constant over the day (modulo TT drift which affects both arms).

The IMC suspension team should double check their filters are on again. I am not familiar with the settings and I don't think they've been added to the SDF.

Attachment 1: RFM_errs.png
RFM_errs.png
Attachment 2: Screenshot_2021-04-29_15-12-56.png
Screenshot_2021-04-29_15-12-56.png
  16099   Thu Apr 29 17:43:16 2021 KojiUpdateCDSRFM

The other day I felt hot at the X end. I wondered if the Xend A/C was off, but the switch right next to the SP table was ON (green light).
I could not confirm if the A/C was actually blowing or not.

  7193   Wed Aug 15 13:24:12 2012 DenUpdateCDSRFM -> OAF

Transmission of signals between RFM and OAF is bad again. Now we do not see any errors in IPC_ERR monitors so models think that they get all data but the data is wrong

oaf.png

  1936   Mon Aug 24 10:43:27 2009 AlbertoOmnistructureComputersRFM Network Failure

This morning I found that all the front end computers down. A failure of the RFM network drove all the computers down.

I was about to restart them all, but it wasn't necessary. After I power cycled and restarted C1SOSVME all the other computers and RFM network came back to their green status on the MEDM screen. After that I just had to reset and then restart C1SUSVME1/2.

  9494   Thu Dec 19 14:40:42 2013 KojiUpdateCDSRFM Time over mitigation for c1mcs

I worked on the mitigation of c1mcs time-over issue this afternoon.

The timing for the c1mcs is successfully reduced from >60us to 45us.


The previous models are svned in redoubt as follows:

MCS rev. 6696
RFM rev. 6697
IOO rev. 6698

What I changed was:

- Remove connection from ALS (on c1ioo) to MCS (on c1sus). This should be all done in LSC. (# of RFM IPC in MCS -1)

- MC2 trans QPD filters are moved from IOO to MCS to reduce the RFM channels in MCS.
  Previously the signals for the 4 segments are sent. Now the processed siganls (pit/yaw/sum) are sent. (# of RFM IPC in IOO -1, MCS -1)

- WFS MC3 feedback channels are moved from MCS to RFM to distribute the RFM channels (# of RFM IPC in MCS -2, in RFM +2)

model    prev. timing[us] current timing[us]  diff in time[us]  diff in ch#
c1mcs         >60                45                -15              -4
c1rfm         47                 53                + 6              +2       
c1ioo         47                 36                -11              -1

Revisions of the new models:
MCS rev. 6702
RFM rev. 6701
IOO rev. 6700

  2195   Fri Nov 6 17:04:01 2009 josephbConfigurationComputersRFM and Megatron

I took the RFM 5565 card dropped off by Jay and installed it into megatron.  It is not very secure, as it was too tall for the slot and could not be locked down.  I did not connect the RFM fibers at this point, so just the card is plugged in.

Unfortunately, on power up, and immediately after the splash screen I get "NMI EVENT!" and "System halted due to fatal NMI". 

The status light on the RFM light remains a steady red as well.  There is a distinct possibility the card is broken in some way.

The card is a VMIPMC-5565 (which is the same as the card used by the ETMY front end machine).  We should get Alex to come in and look at it on Monday, but we may need to get a replacement.

  2486   Fri Jan 8 10:38:35 2010 josephb, kojiUpdateComputersRFM and Megatron

Last night, we installed the VMI 5565 RFM card into Megatron.  After turning off the watchdogs for the ETMY optic, we disconnected the RFM fiber, and connected it to megatron, then powered it up. 

We modified the RCG code to have 3 rfmio blocks, which were reading 0x11a1c0 (ascPit), 0x11a1c4 (ascYaw), and 0x11a1c8 (lscPos).  These were connected to the approriate filter module inputs, and we also added grounds to the front of the rfmio blocks (we looked at the ass code which was setup that way, so we just did the same thing).  When we started it however, it didn't read properly.  If we turned off the input and set and offset, it calculated the output of the filter module correctly, (i.e. just the offset value), but as soon as we turned on the input, it was set to 0, no matter the offset value, which indicated it was reading something correctly.

After this test, the RFM fibers were reconnected to c1iscey, we rebooted c1iscey, and we confirmed that the system was working properly again.  We turned the watch dogs back on for ETMY.

 

 

  2487   Fri Jan 8 11:43:22 2010 josephb, alexUpdateComputersRFM and Megatron

Alex came over with a short RFM cable this morning.  We used it to connect the rfm card in c1iscey to the rfm card megatron

Alex renamed startup.cmd in /cvs/cds/caltech/target/c1iscey/ to startup.cmd.sav, so it doesn't come up automatically.  At the end we moved it back.

Alex used the vxworks command d to look at memory locations on c1iscey.  Such as d 0xf0000000, which is the start of the rfm code location.  So to look at 0x11a1c8 (lscPos) in the rfm memory, he typed "d 0xf011a1c8".  After doing some poking around, we look at the raw tst front end code (in /home/controls/cds/advLigo/src/fe/tst), and realized it was trying to read doubles.  The old rts code uses floats, so the code was reading incorrectly.

As a quick fix, we changed the code to floats for that part.  They looked like:

etmy_lsc = filterModuleD(dsp_ptr,dspCoeff,ETMY_LSC,cdsPciModules.pci_rfm[0]? *(\
(double *)(((void *)cdsPciModules.pci_rfm[0]) + 0x11a1c8)) : 0.0,0);

And we simply changed the double to float in each case.  In addition we changed the RCG scripts locally as well (if we do a update at some point, it'll get overwritten).  The file we updated was /home/controls/cds/advLigo/src/epics/util/lib/RfmIO.pm

Line 57 and Line 84 were changed, with double replaced with float.

return "cdsPciModules.pci_rfm[0]? *((float *)(((void *)cdsPciModules.pci
_rfm[$card_num]) + $rfmAddressString)) : 0.0";

. "  *((float *)(((char *)cdsPciModules.pci_rfm[$card_nu
m]) + $rfmAddressString)) = $::fromExp[0];\n"

This fixed our ability to read the RFM card, which now can read the LSC POS channel, for example.

Unfortunately, when we were putting everything the way it was with RFM fibers and so forth, the c1iscey started to get garbage (all the RFM memory locations were reading ffff).  We eventually removed the VME board, removed the RFM card, looked at it, put the RFM card back in a different slot on the board, and returned c1iscey to the rack.  After this it started working properly.  Its possible in all the plugging and unplugging that the card somehow had become loose.

The next step is to add all the channels that need to be read into the .mdl file, as well as testing and adding the channel which need to be written.

 

  2488   Fri Jan 8 15:40:14 2010 josephb, alexUpdateComputersRFM and RCG

Alex added a new module to the RCG, for generating RFMIO using floats.  This has been commited to CVS.

  4516   Tue Apr 12 16:01:33 2011 josephbUpdateGeneralRFM errors

Problem:

Currently the c1scy, c1mcs, and c1rfm models are reporting an error with receiving some data sent over the GE Fanuc Reflected memory cards.

To be more exact, the C1:SUS-ETMY_ALS signal from the c1gcv FE code on the c1ioo computer going too the Y end is not being received However, the C1:SUS-ETMY_LSC signal is.  So the physical RFM card seems to be working.

Similarly, the TRY signal is being sent correctly from the Y end computer.  The X end is working fine and receiving both LSC and ALS signals.

The c1mcs and c1rfm models also receive data from the c1ioo computer and reporting receiving errors.

Theory:

Because the RFM cards are transmitting and receiving at least some channels, I'm guessing there was changes made to the C1.ipc file, which defines the memory locations of these various channels on the RFM network, and that when a model was rebuilt, a different one using the previous IPC file was not, and thus one of the computer is going to the wrong place to either read or write data.

Tomorrow, I'm planning on the  following:

1) Clean out the C1.ipc file (/opt/rtcds/caltech/c1/chans/ipc/)

2) Rebuild all models

3) Run activate_daq.py script

4) Restart models via script

If this doesn't clear up the problem, I'll continue  to bug hunt.

  14293   Tue Nov 13 21:53:19 2018 gautamUpdateCDSRFM errors

This problem resurfaced, which I noticed when I couldn't get the single arm locks going.

The fix was NOT restarting the c1rfm model, which just brought the misery of all vertex FEs crashing and the usual dance to get everything back.

Restarting the sender models (i.e. c1scx and c1scy) seems to have done the trick though.

Attachment 1: RFMerrors.png
RFMerrors.png