  6335   Tue Feb 28 16:44:56 2012 ranaUpdateLSCMICH and PRCL signals in a simulation

 

 Like I said, this is possible if you fail to set up the OSA to look at the sidebands at BOTH the AS and REFL ports at all times. Don't waste your time - set up an OSA permanently!

  9118   Mon Sep 9 20:46:28 2013 MasayukiUpdateLSCMICH calbration

[Manasa, Masayuki]

We took a bunch of measurements. Transfer function and power spectrum using DTT. They will be used to obtain calibrated MICH in-loop and free-running noise. Detail Elog with plots will follow very soon.

  9121   Tue Sep 10 17:35:50 2013 Masayuki, ManasaUpdateLSCMICH calbration

Quote:

[Manasa, Masayuki]

We took a bunch of measurements. Transfer function and power spectrum using DTT. They will be used to obtain calibrated MICH in-loop and free-running noise. Detail Elog with plots will follow very soon.

 [Masayuki, Manasa]

Estimation of free-running MICH displacement noise:

Method 1. Assuming AS55_Q_err to be a linear sensor, as shown in (1) of figure below, free-running MICH noise (V_d) can be estimated by measuring V_err and the OLTF G. H can be estimated by using method explained in elog

 Method 2. Considering that the AS55_Q signal might be distorted or saturated, method 1 may not be precise. In method 2, we will use the ASDC as the sensor (S' in (3)) instead and lock MICH using ASDC in mid-fringe to calibrate the ITM actuators.

Figure:1

Schematic:

MICH_calib_loops.png

What we did:

1. Estimate H' from free-running ASDC signal (bright to dark fringe).
2. With MICH locked on ASDC, give an excitation signal to C1:LSC-SUS_XXXX_EXC (XXXX could be ITMX or ITMY) and measure R'. [(3) of schematic]
3. Measure OLTF of MICH locked on ASDC (hence estimate L). [(3) of schematic]
4. With MICH locked on AS55_Q, give an excitation signal to C1:LSC-SUS_XXXX_EXC (XXXX could be ITMX or ITMY) and measure R1. [(2) of the schematic] 

Results/Plots: 

Figure:2

OLTF of MICH locked on ASDC

 OLTF_MICHDC.png

 

Figure2:

Actuator excitation to MICH transfer function (MICH locked using ASDC) 

MICH_DC_resp.png

* y axis (no units)

Figure 3:
Actuator excitation to MICH transfer function (MICH locked using AS55Q)

MICH_RF_resp.png 

* y axis (no units)

Figure 4:
Free-running MICH noise

MICH_free_noise.png 

Discussion: 

1. By using the second sensor, we also eliminate the effect of the MICH servo loop locked on AS55_Q (Estimated V_d does not depend on G but only on G').

2. The free-running MICH noise is still suppressed at 1Hz. This should be coming from the effect of the UGF of the loop at ~10Hz and the vicinity to the pendulum frequency at 1Hz.

 

Edit/Masayuki// This noise curve is not collect, especially in low frequency region. We used the measured OLTF for compensating the free running noise, but that is not collect in low frequency region. So we should model the OLTF and fit that into the measured OLTF. We will fix this soon.

 

 

  9127   Thu Sep 12 23:36:25 2013 MasayukiUpdateLSCMICH calbration

For Modelling of the OLTF, I measured the response of the BS suspension. I used the OSEM sensor for measurement. The attatchment1 is the measured TF from C1:SUS-BS_LSC_EXC to C1:SUS-BS_SUSPOS_IN1 with exciting with random force. The measured data was fitted and the resonant frequency is 1.029(±0.005) Hz and quality factor is 12.25 (± 0.2).  Additionally I did same measurement for ITMX and ITMY. The attachment 2 and 3 are the results for ITMX and ITMY. Each eigenfrequency and Q are 1.063 (±0.008) Hz and 7.33 (±0.13) (ITMX), 1.022 (±0.005) Hz and 9.41 (±0.09) (ITMY).

 After that, I locked the MICH with AS55, and measured the PSD of error signal. I compensated the that PSD by the modelled OLTF with this suspension TF and the servo TF. The result is in attachment 4. Above 1 Hz it is quite close to the previous data by Keiko (elog#6385) But below 1 Hz there is a large dip. The error signal has also this dip. I looked for a integral filter between 0.2 Hz and 1 Hz, but I connot find a such filter. And when I locked MICH with using ASDC, there was same dip at same frequency. I don't think it's true free running noise, and I will try to fix it.

I completely forgot to mention that I fitted the modelled OLTF into the measured OLTF. I used the fitted OLTF for compensation. 

 

 

Attachment 1: BSsus.PNG
BSsus.PNG
Attachment 2: ITMXsus.PNG
ITMXsus.PNG
Attachment 3: ITMY.PNG
ITMY.PNG
Attachment 4: free_running.PNG
free_running.PNG
  9128   Fri Sep 13 19:22:01 2013 MasayukiUpdateLSCMICH calbration

 

 I made sure the yesterday's result was collect. I measured not only the error signal but also the feedback signal. And I compared those signals and measured the TF in order to confirm my servo filter model is not wrong.

The reason of dip at low frequency region is maybe the coherence of the ground motion. The ITMX and ITMY suspensions are put close. If ground motion has coherence, the mirrors move in common mode. That will suppress the free running noise. The attachment is the free running noise of Sep 13rd and Sep 12nd.

Attachment 1: noise.PNG
noise.PNG
  9131   Mon Sep 16 14:11:47 2013 ranaUpdateLSCMICH calbration

  There doesn't seem to be any coherence among the different directions of ground motion (as expected from seismic theory), so I am suspicious of such a low MICH noise.

Attachment 1: Screen_Shot_2013-09-16_at_2.10.31_PM.png
Screen_Shot_2013-09-16_at_2.10.31_PM.png
Attachment 2: Screen_Shot_2013-09-16_at_2.18.47_PM.png
Screen_Shot_2013-09-16_at_2.18.47_PM.png
  9134   Tue Sep 17 00:50:42 2013 MasayukiUpdateLSCMICH calbration

I found the bug in my calibration code, and I fixed it.

And I put the white Gaussian noise on the BS actuator, and calibrated to the differential length with my new code. We already know the efficiency of the actuator(elog#8242), so I could estimate how much I put the disturbance and compare the two values. The result is in attachment 1.  x_exc means the value of the disturbance. 

You can see the PSD of the differential motion decrease factor of 3 by decreasing the disturbance by factor of 3 (except for the region from 1 Hz to 5 Hz), and the value at lower frequency than resonant frequency of the suspension is comparable to the value estimated with the actuator efficiency. Also there is no dip when I put the larger disturbance than free running noise.

Between 1 Hz and 5 Hz there seems to be a resonance of something (seismic stack?). And also on resonance of the suspension there seems to be some other noise source. One possibility is the active damping of each suspension.

Actually still there seems to be a dip between 0.1 Hz and 1 Hz. But if you consider about those effect, I think this result doesn't seems to be so strange. But according to the documentation of LIGO document-T000058, which I found the seismic motion in 40 m Lab is written in, the seismic motion at 0.1 Hz is 10^-7. I'm not sure about this factor of 10 difference. One possibility is the geophone doesn't have good sensitivity at low frequency. I'm still not sure this result is really collect.

 


 

Attachment 1: noise.PNG
noise.PNG
  6362   Tue Mar 6 01:35:03 2012 kiwamuUpdateLSCMICH characterization

[Keiko / Kiwamu]

 Update on the MICH characterization:

  • The OSAs weren't so great because the 11 MHz sidebands were covered by the carrier's tail
    • It seemed that the frequency resolution depended on the mode matching. We will try improving the mode matching tomorrow.
  • The noise budget looked very bad
    • There were huge peaks at 1 Hz, 3 Hz, 16.5 Hz and 23 Hz. Something is crazy in the vertex suspensions.
    • Keiko will post the calibrated noise budget.
  • The MICH response at AS55Q was measured and we will calibrate it into watts / meter.

 

  10803   Tue Dec 16 01:50:27 2014 rana, diegoFrogsLSCMICH filter is nuts

 This is ridiculous.

How many RGs can I fit into one button???

Attachment 1: badMICHrg.pdf
badMICHrg.pdf
  9293   Fri Oct 25 20:11:08 2013 JenneUpdateLSCMICH gain in PRMI lowered by factor of 2

We were locking the PRMI, but it is very rumbly today.  I reduced the MICH servo gain from -0.8 to -0.4 , and things seem to be better.  Now my MICH UGF is about 60Hz.

  9809   Mon Apr 14 19:02:09 2014 JenneUpdateLSCMICH gets noisy as CARM or DARM offset reduced

This afternoon, I was toying around with reducing either the CARM or DARM offsets (so, put in a CARM offset, leave DARM zero, lock the PRMI, then reduce CARM offset to zero.  Or, put in a DARM offset, leaving CARM offset zero, lock the PRMI, then reduce the DARM offset to zero).

When looking at the data, I see that the MICH error signal gets fuzzier when the arms get close to resonance. (Note here that because I forgot to zero the carm offset before finding the resonances, -3 is my zero point for this plot and the next.)

MICH_fuzzy_when_offsets_small_longerData.png

Here is a zoom of the last piece of this time series, but with both TRX and TRY plotted (along with POPDC, CARM_ERR and DARM_ERR), where you can see that I had a momentary power buildup of > 100 transmission counts, which is about 20% of our final expected power.

TRX_TRY_100cts.png

Here is a different time series, showing a reduction of the DARM offset, and you can see that as the offset approaches zero, the MICH error signal gets noticeably more fuzzy.  Somewhere near the 240 second mark, I lose PRMI lock.

MICH_fuzzy_when_offsets_small.png

  9813   Tue Apr 15 09:32:19 2014 GabrieleUpdateLSCMICH gets noisy as CARM or DARM offset reduced

I guess this is normal. DARM has (almost) the same effect of MICH on the corner interferometer signals, just increased in amplitude by the arm cavity amplification. When the arm is out of resonance, DARM is almost completely depressed and it is not affecting MICH at all. On the other hand, when the cavities are exactly at resonance, DARM signal is amplified w.r.t. MICH by the cavity gain (2F/pi).

Since DARM is still controlled with ALS, it is probably noisy. The closer to resonance you move the cavities, the more ALS noise in DARM will affect MICH.


Quote:

When looking at the data, I see that the MICH error signal gets fuzzier when the arms get close to resonance. (Note here that because I forgot to zero the carm offset before finding the resonances, -3 is my zero point for this plot and the next.) 

  5413   Thu Sep 15 01:17:10 2011 KeikoUpdateLSCMICH locked and attempt to lock PRCL

 Anamaria, Keiko

- We aligned MICH and were successfully locked MICH using AS55Q. The other mirrors were misaligned so that the other degrees of freedom didn't exist. AS55 was fed back to BS. The f2a filters on BS suspension were required to lock, because the pos feedback was unbalanced to angle degrees of freedom.

- We tried to lock PRCL next, however, because we aligned the MICH and the REFL beam paths were changed, REFL PDs didn't have the light anymore. The REFL paths were modified now, so we will try the PRCL locking next.

- We couldn't confirm REFL55 signals although we alined the REFL paths to REFL55 PD.

  5414   Thu Sep 15 02:18:19 2011 AnamariaUpdateLSCMICH locked and attempt to lock PRCL

Kiwamu, Keiko, Anamaria

 

We were able to lock PRC using REFL11I after improving the MICH dark fringe a bit (moving BS) and rotating AS55 and REFL11 such that the signal was maximized in the phases we were using. The dark port is not so dark... but the lock is stable.

I had finished the whole REFL path alignment, but I didn't have a good input beam reference at the time, which is why we had to realign the PDs and the camera. We only had strength to realign 11 and 55. Otherwise, we just need to tweak and center beam on 33 and 165, figure out what's up with 55 and be done with the AP table mods. I hope.

 

Quote:

 Anamaria, Keiko

- We aligned MICH and were successfully locked MICH using AS55Q. The other mirrors were misaligned so that the other degrees of freedom didn't exist. AS55 was fed back to BS. The f2a filters on BS suspension were required to lock, because the pos feedback was unbalanced to angle degrees of freedom.

- We tried to lock PRCL next, however, because we aligned the MICH and the REFL beam paths were changed, REFL PDs didn't have the light anymore. The REFL paths were modified now, so we will try the PRCL locking next.

- We couldn't confirm REFL55 signals although we alined the REFL paths to REFL55 PD.

 

  17270   Tue Nov 15 19:00:56 2022 yutaSummaryBHDMICH locked with balanced homodyne readout at some LO phase

[Paco, Yuta]

MICH was locked with BHD DCPD A-B signal with LO phase controlled.
Locking procedure and configuration was as follows (see Attachment #1).

1. Lock MICH with AS55_Q, with C1:LSC-MICH_GAIN=-3, FM4, FM5, FM8, FM10 (boost filters are turned off to have more phase margin).

2. Lock LO PHASE with BH55_Q, with C1:HPC-LO_PHASE_GAIN=6, FM5, FM8, feeding back to AS1.
  - C1:LSC-BH55_PHASE_R=136.136 deg was tuned to minimize I when AS-LO is fringing with MICH locked with an offset of 50 (we first thought 136.136 deg - 90 deg is better from 40m/17216, but today, 136.136 deg seems to work better; Reason needs to be investigated).
  - We are supposed to use C1:HPC-BH55_Q_AS1_DEMOD_I_OUT to control the LO phase to give maximum MICH signal on BHD_DIFF (40m/17170), but somehow BH55_Q without audio dither was OK to get MICH signal. Line injection at 211.1 Hz on BS was seen in BHDC_DIFF (and AS55_Q), even if we use BH55_Q to lock LO PHASE (see Attachment #2; MICH_B is BHDC_DIFF and MICH_A is AS55_Q) or BH55_Q_AS1_DEMOD_I to lock LO PHASE (with both signs). Reason needs to be investigated.
  - Audio dither was done using AS1 with excitation of 15000 counts at 281.79 Hz. C1:HPC-BH55_Q_AS1_DEMOD_PHASE=60 deg was tuned to minimize Q with injection of line at 13 Hz using LO1.

3. Handed over MICH lock from AS55_Q to 0.66 * C1:LSC-DCPD_A - 1 * C1:LSC-DCPD_B. This was done by using C1:LSC-MICH_A and MICH_B gains. C1:LSC-MICH_A_GAIN=1 was handed over to C1:LSC-MICH_B_GAIN=-1.
  - 0.66 * A - B was tuned so that BHDC_DIFF will be zero (as it supposed to be with MICH offset of zero).
  - AS55_Q and BHDC_DIFF had roughly the same optical gain at 211.1 Hz (actually, BHDC_DIFF had higher optical gain; see Attachment #2), so we used MICH_A_GAIN=1 and C1:LSC-MICH_B_GAIN=-1
  - After handing over of BHDC_DIFF, OLTF was measured. UGF was ~70 Hz (Attachment #3).

Next:
  - Investigate how to get optimal LO phase. With BH55_Q or BH55_Q + audio dithering? How to optimize demod phases?
  - How do we balance DCPD A and B? What is the effect of BHD BS being 44:56 not 50:50?
  - Measure amount of MICH signal in BHDC_DIFF with different LO phases.
  - Improve SNR in BH55.
  - It will be much simpler if we send BHDC_SUM and BHDC_DIFF to c1lsc from c1hpc, instead of sending un-unwhitened BHDC_A and B.

Attachment 1: Screenshot_2022-11-15_18-34-25_MICHBHDLocked.png
Screenshot_2022-11-15_18-34-25_MICHBHDLocked.png
Attachment 2: Screenshot_2022-11-15_19-02-02_MICHLine.png
Screenshot_2022-11-15_19-02-02_MICHLine.png
Attachment 3: Screenshot_2022-11-15_18-38-20_BHDMICHOLTF.png
Screenshot_2022-11-15_18-38-20_BHDMICHOLTF.png
  9082   Wed Aug 28 07:33:51 2013 manasaUpdateLSCMICH locking

I wanted to measure the OLTF of MICH.

What I did:
1. Ran LSC offsets script to zero all the offsets.

2. Restored the IFO configure settings for locking Michelson (locked on AS55Q).

3. MICH wouldn't lock on these settings.

4. The MICH servo was hitting its limits (10000 counts). I checked the filter module. After a little bit of looking into things, I disabled FM3 (0,0:5,5), FM4 (1:10) and FM7 (1:5). FM3 and FM7 were filter modules that were switched ON at the trigger. I set these to manual. Enabling any of the filters (FM3, FM4, FM7) caused MICH to lose lock.

5. MICH gain was changed from -20 to -30. MICH locked with ASDC suppressed to 0.01 counts. I looked at the power spectrum of C1:LSC-MICH_OUT on dtt. //edit: Manasa// The plot (uncalibrated) now shows MICH_OUT power spectrum with MICH PSL shutter closed, free-running MICH and loop-enabled MICH.

6. I then wanted to measure the OLTF of MICH using dtt. A channels were set to C1:LSC-MICH_IN1/C1:LSC-MICH_IN2 and excitation given through C1:LSC-MICH_EXC. But I have not been able to get any good coherence for the measurement as yet.

Attachment 1: MICH.pdf
MICH.pdf
  16228   Tue Jun 29 17:42:06 2021 Anchal, Paco, GautamSummaryLSCMICH locking tutorial with Gautam

Today we went through LSC locking mechanics with Gautam and as a "Hello World" example, worked on locking michelson cavity.


MICH settings changed:

  • Gautam at some point added 9 dB attenuation filters in MICH filter module in LSC to match the 9 dB pre-amplifier before digitization.
  • This required changing teh trigger thresholds, C1:LSC-MICH_TRIG_THRESH_ON and C1:LSC-MICH_TRIG_THRESH_OFF.
  • We looked at C1:LSC-AS55_Q_ERR_DQ and C1:LSC-ASDC_OUT_DQ on ndscope.
  • The zero crossings in AS55_Q correspond to ASDC going to zero. We found the threshold values of ASDC by finding the linear region in zero crossing of AS55_Q.
  • We changed the thresold values to UP: -0.3mW and DOWN -0.05mW. The thresholds were also changed in C1LSC_FM_TRIG.
  • We also set FM2,3,6 and 8 to be triggered on threshold.

We characterized the loop OLTF, found the UGF to be 90 Hz and measured the noise at error and control points.

gautam: one aim of this work was to demonstrate that the "Lock Michelson (dark)" script call from the IFOconfigure screen worked - it did, reliably, after the setting changes mentioned above.

  16889   Fri Jun 3 17:42:50 2022 yutaUpdateLSCMICH locks with AS55_Q

AS path at AP table as re-aligned and confirmed that MICH can be locked with AS55 Q.

What we did:
  - Aligned AS55 and AS110 paths at AP table. AS55 was not receiving enough light. AS110 was not receiving light at all.
  - Changed AS55 I and Q whitening gain from 3dB to 42dB.
  - Zero-ed the RF offsets manually. C1:LSC-AS55_Q_IN1 is having too large offset. When PSL shutter was closed, it reads 13950! Needs investigation.
  - Locked MICH with PRM mis-aligned with configurations attached.

Other Issues:
 - C1:IOO-MC_TRANS_SUM is now stuck at 14009. MC auto locker doesn't work correctly. FIX ME!

Attachment 1: Screenshot_2022-06-03_17-41-55.png
Screenshot_2022-06-03_17-41-55.png
  15000   Wed Oct 30 11:53:41 2019 gautamUpdateLSCMICH loop shape tuning

I changed the shape of the low pass filter to reduce high frequency sensor noise injection into the MICH control signal. The loop stability isn't adversely affected (lost ~5 degrees of phase margin but still have ~50 degrees), while the control signal RMS is reduced by ~x10. This test was done with the PRMI locked on the carrier, need to confirm that this works with the arms controlled on ALS and PRMI lcoked on sideband.

Attachment 1: MICH_ELP.pdf
MICH_ELP.pdf
Attachment 2: MICH_ELP_TFs.pdf
MICH_ELP_TFs.pdf
  1330   Fri Feb 20 19:31:16 2009 YoichiUpdateLSCMICH low gain problem
Last night, we found that MICH UGF was too low. Even after re-aligning the PDs, it was still too low.
Today, I compared the UGFs of MICH and PRC when in the DRMI configuration locked with the single demod. signals.
In this configuration, MICH signal comes from REFL33Q and the PRC signal comes from REFL33I (the same PD).
The PRC UGF was about 100Hz whereas MICH was only ~10Hz.
Since they uses the same PD, the low gain is not caused by the PD.
I checked conlog history and confirmed there is no change in the MICH->BS path in the last few days.
I also checked the svn history of chans directory for changes in filters. Nothing problematic found.

Then I noticed that the susvme computers were overloaded.
This time, I rebooted all the FE computers just in case.

Then the MICH gain was somewhat recovered (by a factor of 3 or so). Don't know why.

I adjusted the DD_handoff script to set the MICH gain to 0.7 before the bounce-roll filter is engaged.
  6368   Tue Mar 6 23:37:31 2012 keikoUpdateSUSMICH noise budget - SUS check

Here are the OSEM spectrum of MICH suspensions (BS, IX, IY). Bounce and Roll modes are shown on 16 and 23 Hz. The filters for them has been checked.

Mar6sus1.pdf

Mar6sus2.pdf

Mar6sus3.pdf

keiko, kiwamu, Rana

Attachment 1: Mar6sus1
Attachment 2: Mar6sus2
Attachment 3: Mar6sus3
  6376   Wed Mar 7 17:39:40 2012 keikoUpdateLSCMICH noise budget on 5 Mar

 This is the calibrated MICH noise budget on Mar 5. There was a sharp peak at 1Hz and a blob on 3 Hz. The demod phase was adjusted for AS55Q.

Mar5-MICHbudget.png

 

Attachment 1: Mar5-MICHbudget.png
Mar5-MICHbudget.png
  6380   Wed Mar 7 20:53:13 2012 keikoUpdateLSCMICH noise budget on 5 Mar

 

 Mar6-MICHbudget.png

This is the MICH noise budget on 6th March. 1Hz peak got a bit better as the BS sus control gain was increased.

 

  6385   Thu Mar 8 00:57:48 2012 keikoUpdateLSCMICH noise budget on Mar 5, Mar 6, and old

Here is the recent two noise budgets of MICH, with the old measurement by Jenne. The most latest Mar 6 data is quite close to the old data, even better around 20-30 Hz. Probably some scattering source was improved?

Mar7MICHbudgettotal.png

  4822   Wed Jun 15 02:20:00 2011 JenneUpdateLockingMICH noise budget?

I would like to announce my confusion with regard to the MICH noise budget, in hopes that someone else has some inspiration

If you tilt your head sideways, you will notice that in this plot (totally uncalibrated, as yet), the BLACK trace, which is my white-light measurement of the AS55 shot noise is above the AS55Q noise when the Michelson is locked (true only at low frequency).  You will also notice that the same appears to be true for the Whitening Filter + Antialiasing Filter + ADC noise (GRAY trace).  Since Black, Gray, Pink and Green should all have the same calibration factor (a constant), calibrating the plot will not change this.  Brown and Blue are the MICH_OUT (aka MICH_CTRL) for dark and bright fringes, respectively.

I measure 58mV at the DC out of the AS55 PD when the Michelson is locked on the bright fringe.  This (assuming DC transimpedance of 50ohms) gives 1.16mA of DC photo current.

So.  What is going on here?  Am I totally confused??

In other news, assuming (which I'm not 100% confident about right now) that these traces are vaguely correct, the Michelson is limited by shot noise above ~20Hz.  This is...good?  We want to be shot noise limited.  Do we want to be limited at such a low frequency?

(Also, yes I can calibrate the plot to m/rtHz, but no, I won't tonight because something is funny with my calibration for the free running noise and I'll fix it tomorrow.)

MICH_noise_budget_measurements_15June2011.jpg

  4823   Wed Jun 15 12:16:37 2011 JamieUpdateLockingMICH noise budget?

Quote
If you tilt your head sideways, you will notice that in this plot (totally uncalibrated, as yet), the BLACK trace, which is my white-light measurement of the AS55 shot noise is above the AS55Q noise when the Michelson is locked (true only at low frequency).  You will also notice that the same appears to be true for the Whitening Filter + Antialiasing Filter + ADC noise (GRAY trace).  Since Black, Gray, Pink and Green should all have the same calibration factor (a constant), calibrating the plot will not change this.  Brown and Blue are the MICH_OUT (aka MICH_CTRL) for dark and bright fringes, respectively.

Hey, Jenne.  I think there are a couple of things.  First, you're missing a PD dark noise measurement, which would be useful to see.

But I think the main issue is that it sounds like all of your closed loop measurements are done with the in-loop PD.  This means that everything will be suppressed by the loop gain, which will make things look like they have a noise lower than the actual noise floor.

  17282   Thu Nov 17 20:02:10 2022 yutaSummaryBHDMICH optical gain measurements with different LO phases

MICH optical gain was measured with different LO phases over ~90 degrees.
Zero crossing of BH55_Q_ERR seems to be roughly 55 degrees away from optimal LO phase.

What we did:
 - Locked MICH with AS55_Q with no offset, with UGF at ~10 Hz (same as configuration in 40m/17279).
 - Injected BS calibration line at amptilude of 300 counts at 211.1 Hz.
 - Locked LO Phase with BH55_Q with different offsets added at C1:HPC-LO_PHASE_OFFSET.
 - Measured sensing matrix at that frequency. Counts are calibrated into meters using actuator efficiencies as described in 40m/17279.
 - LO phase was obtained using a DC value of BH55_Q. This was calibrated into degrees from the following:

Amplitude of LO-AS fringe in BH55_Q was calculated to be

A = BH55optgain*lamb/(4*pi) = 60 counts

where BH55optgain is 7.12e8 counts/m, which is optical gain of BH55_Q for LO1 measured in 40m/17279.
(Actually, BH55_Q goes upto ~ +/-200 counts in time series data, but maybe 60 is the nominal fringe amplitude, considering alignment fluctuations and fluctuation in AS darkness? Note that, no offset in BH55_Q is assumed in this calculation, but AM etc can create an offset.) 
LO phase can be obtained by

LOphase = arcsin(BH55_Q/A)

where BH55_Q a DC value (10 sec average) of BH55_Q.

Result:
 Attachment #1 is uncalibrated plot C1:HPC-LO_PHASE_OFFSET of around +/- 50 was the maximum we could add, and more offset gave unstable lock.
 Attachment #2 is calibrated plot. AS55_Q does not depend on LO phase, as expected. BH55_Q and BHDC_DIFF depend on LO phase as expected. BH55_I and AS55_I stay at low level, as expected (this means that our RF demodulation phase is OK).
 Dotted gray line is an eyeball fit of expected curve (40m/17170) to fool your eyes.
 This tells you that we are roughly 55 deg away from LO phase which gives maximum MICH signal for BHDC_DIFF.
 Error bar in x-axis is from standard deviation of BH55_Q fluctuations. Error in y axis is probably ~20% at maximum.
 Notebook: /opt/rtcds/caltech/c1/Git/40m/scripts/CAL/SensingMatrix/MeasureSensMatBHD.ipynb

Next:
 - Repeat the measurement with
   - MICH locked with higher UGF, with notch at 211.1 Hz, for more robust AS dark fringe
   - DCPD A and B balanced at 211.1 Hz (null MICH signal for BHDC_SUM to balance?)
   - Measure optical gain also for BHDC_SUM and BH55_Q demodulated at audio dither
   - Lock LO phase at different sign so that we can sweep LO phase over ~180 deg
   - Sign-sensitive optical gain measurement (demodulation with BS motion necessary)
 - Compare with expected values from simulations
 - Why do we have 55 degrees offset? Expected offset is 90 degrees...
   - Check if there is any RAM in 55 MHz in the input beam by measuring AM with ITM single bounce

Attachment 1: UncalibratedMICHOpticalGainVSLOphase20221117.png
UncalibratedMICHOpticalGainVSLOphase20221117.png
Attachment 2: CalibratedMICHOpticalGainVSLOphase20221117.png
CalibratedMICHOpticalGainVSLOphase20221117.png
  17287   Fri Nov 18 22:46:02 2022 yutaSummaryBHDMICH optical gain measurements with different LO phases, with signs

MICH optical gain with a sign was measured with different LO phases over ~180 degrees, with updated calibration and higher MICH UGF.
Zero crossing of BH55_Q_ERR seems to be 68 degrees away from optimal LO phase.

Calibrated sensing matrix:
 - Locked MICH with AS55_Q at dark fringe, with UGF of ~200 Hz. Notch at 311.1 Hz was turned on.
 - Locked LO PHASE with BH55_Q, with UGF of ~10 Hz (C1:HPC-LO_PHASE_GAIN=-2, using LO1).
 - Measured the sensing matrix as written in 40m/17279, but with different dither frequencies to avoid violin mode frequencies and to match with already-installed notch filters.
 - Sensing matrix was calibrated into meters using actuator gains measured in 40m/17285
 - Sign was added by comparing the phase with C1:SUS-xx_LSC_OUT. If they are 90-270 deg apart, minus sign was added to the sensing matrix.
 - Resuts are as follows. At least important green ones are consistent with previous measurements (40m/17279).

Calibrated sensing matrix with the following demodulation phases (counts/m)
{'AS55': -164.1726747789845, 'BH55': 169.57651332419115}
      Sensors            MICH @311.1 Hz           LO1 @147.1 Hz           AS1 @141.79 Hz           
C1:LSC-AS55_I_ERR_DQ    2.72e+06 (84.89 deg)    6.41e+05 (14.60 deg)    -1.98e+05 (206.79 deg)    
C1:LSC-AS55_Q_ERR_DQ    -1.20e+09 (-228.85 deg)    -1.43e+06 (-106.51 deg)    1.41e+06 (29.21 deg)    
C1:LSC-BH55_I_ERR_DQ    -2.45e+09 (-230.64 deg)    -6.57e+07 (167.84 deg)    7.28e+07 (-16.56 deg)    
C1:LSC-BH55_Q_ERR_DQ    7.81e+09 (-48.64 deg)    -7.34e+08 (159.70 deg)    8.06e+08 (-10.08 deg)    
C1:HPC-BHDC_DIFF_OUT    -9.91e+08 (-224.55 deg)    -1.13e+08 (164.14 deg)    1.26e+08 (-3.95 deg)    
C1:HPC-BHDC_SUM_OUT    -6.84e+06 (-104.69 deg)    1.50e+07 (-8.71 deg)    -1.79e+07 (173.80 deg)    
LO phase from C1:LSC-BH55_Q_ERR_avg 4.98e-03 +/- 1.65e+01 deg

Estimating LO phase:
 - Using 7.34e+08 counts/m, which is an optical gain of BH55_Q for LO1, LO phase can be estimated as follows.

A = BH55optgain*lamb/(4*pi) = 62 counts
LOphase = arcsin(BH55_Q/A)

 - When C1:HPC-LO_PHASE_GAIN is plus, LOphase was calculated with the following to take into account of the sign flip in the controls.

LOphase = 180 - arcsin(BH55_Q/A)

Balancing A-B:
 - BHDC_A and BHDC_B were balanced to give null MICH signal in BHDC_SUM at 311.1 Hz. This gave BHDC_DIFF = 0.919*A - B.
 - It seems like this balancing gain changes over time by ~30%.

Result:
 - Attachment #1 is uncalibrated MICH optical gain in different LO phases, and Attachment #2 is the calbirated one. Basically the same with 40m/17282, but with updated calibration and sign considerations.
 - In addition to the previous measurements, we can see that BHD_SUM is not dependent on LO phase (small dependence probably from not perfect A and B balancing).
 - 0 deg of LO phase means that it is a zero crossing of BH55_Q with a slope that LO PHASE loop can be closed with a minus C1:HPC-LO_PHASE_GAIN, feeding back to LO1.
 - Dotted and dashed gray lines are from scipy.optimize.curve_fit using the following fitting function (not an eyeball fit this time!).

def fitfunc(x, a,b,c):
    return a*np.sin(np.deg2rad(x-b))+c

 - Fitting results show that we are -22 deg away from our intuition that BH55_Q crosses zero when BHDC_DIFF give no MICH signal (68 degrees away from optimal LO phase).
 - Fitting results also show that BH55_Q sensitivity to MICH crosses zero when BHD_DIFF sensitivity to MICH maximizes. This suggests that BH55+MICH dither can be used to lock LO phase to optimal LO phase.

Notebook: /opt/rtcds/caltech/c1/Git/40m/scripts/CAL/SensingMatrix/MeasureSensMatBHD.ipynb

Next:
 - Compare with expected values from simulations
 - Why do we have -22 deg?
   - Check if there is any RAM in 55 MHz in the input beam by measuring AM with ITM single bounce (quick measurement shows it is small)
   - Unbalanced BHD BS?
   - Contribution from 55 MHz sidebands from LO beating with 55 MHz sidebands from AS?
       - Lock LO phase using audio dither only (demodulate BHDC_DIFF?).

Attachment 1: UncalibratedMICHOpticalGainVSLOphase20221118.png
UncalibratedMICHOpticalGainVSLOphase20221118.png
Attachment 2: CalibratedMICHOpticalGainVSLOphase20221118.png
CalibratedMICHOpticalGainVSLOphase20221118.png
  9478   Mon Dec 16 02:20:49 2013 DenUpdateLSCMICH rms is improved

When PRMI is locked on REFL 165 I&Q signals MICH rms is dominated by the 60 Hz line and harmonics. It comes from demodulation board.

To increase SNR ZFL-100 LN amplifier (+23.5dB) was installed in LSC analog rack. MICH 60 Hz and harmonics are improved as shown on the plot "mich_err"

I have also added a few resg at low frequencies. MICH rms is not 3*10-10. In Optickle I simulated power dependence in PRC and ARMs on MICH motion. Plot is attached.

 I think we need to stabilize MICH even more, down to ~3*10-11 . We can think about increasing RF amplifier gain, modulation index and power on BB PD.

CARM offset reduction was a little better today due to improved MICH RMS. Power in arms increases up to 15 and than starts to oscillate up to 70 and then PRMI looses lock.

Tomorrow we need to discuss where to put RF amplifier. Current design has several drawbacks:

  • DC power for the amplifier is wired from a custom (not rack based) +15V power supply that was already inside the lsc rack and used for other ZFL-100LN
  • BNC cables are used because I could not find any long SMA cables
  • we would like gain of ~40 dB instead of 23.5 dB
Attachment 1: MICH_ERR.pdf
MICH_ERR.pdf
Attachment 2: DC_power.pdf
DC_power.pdf
Attachment 3: ARM_OFFSET.pdf
ARM_OFFSET.pdf
  9767   Mon Mar 31 17:47:57 2014 ericqSummaryLSCMICH sensing oddities in REFL 3F

Last week, while I had the PRMI locked on REFL33, I did some poking around with mirror excitation to RFPD quadrature transfer functions. I got some indication of weird things with sensing MICH with the 3F REFL signals, but it should be explored more before taken as a real thing. I just figured I would show what I saw. 

With that disclaimer out of the way, here's what I did:

  • Locked PRMI on PRCL:REFL33_I and MICH:REFL33_Q, as detailed in my earlier ELOG
  • Created PRCL and MICH excitations at two different frequencies, notched said frequencies out of the control filters
  • Took transfer functions from mirror LSC output signals to 33 I, 33 Q, 165 I, 165 Q in DTT
  • For each DOF, look at the measured transfer functions only at the excitation frequency. (Assuming good coherence, which was there)

The basic idea was, some PRCL motion (for instance), has a transfer function to both the I and Q quadratures at a given PD. As the PRCL excitation sine wave goes through one cycle, the REFL signals at the excitation frequency go through some coherent cycle. Thus, the excitation traces out some trajectory in the I vs. Q plane. I believe this is analogous to the typical "radar plot" that we make for sensing matrix elements. 

However, the straight line that we normally plot in the radar plots assumes a certain phase relationship between the DOF-> I and DOF->Q transfer functions that results in a straight line. Here are the trajectories I actually measured, normalized by the excitation amplitudes.

REFL_33_traj.pdfREFL_165_traj.pdf

The plotted traces are (x,y) = (H_prcl->I * prcl, H_prcl->Q * prcl) and  (x,y) = (H_mich->I * mich, H_mich->Q * mich) where H_prcl->I is the measured complex transfer function from prcl to REFL I, for instance, and prcl and mich are the excitation signals, normalized to unit amplitude.

PRCL looks like a nice straight line in both of these, and pretty well phased, but not only is MICH not very orthogonal to PRCL, there is quite a bit of ellipticity present, which means we can't fully decouple the two DOFs, even if they were nominally orthogonal. 

I'm not sure what may cause this. To back up this measurement/interpretation, I tried to take measurements of these transfer functions across different excitation frequencies via swept sine DTT, but seismic activity kept me from staying locked long enough...

  9768   Mon Mar 31 21:23:30 2014 GabrieleSummaryLSCMICH sensing oddities in REFL 3F

I'm not sure what may cause this. To back up this measurement/interpretation, I tried to take measurements of these transfer functions across different excitation frequencies via swept sine DTT, but seismic activity kept me from staying locked long enough...

I guess that you get an ellipse when the transfer functions to I and Q have a different phase. One mechanism could be that when driving MICH we make some residual PRCL and this couples with a different transfer function to both I and Q. However, I would expect no phase lag in the PRMI configuration, since there is no enough optical delay in the system to give significant dephasing at few hundreds Hz. This effect might come from mechanical resonances.

It is worth measuring the optical transfer functions from both PRCL and MICH to REFL signals at all frequencies, to see if we have strange features in the TFs.

  13395   Thu Oct 19 15:42:03 2017 jamieSummaryLSCMICH/PRCL reconstruction neural network running on c1lsc

Gabriele's PRCL/MICH reconstruction neural network is now running on c1lsc.  Summary:

  • front-end model is called c1dnn, and is running as an experimental user-space process
  • c1dnn is getting most of it's needed inputs from existing SHMEM IPC outputs from c1lsc
  • none of the output signals from the network are being sent anywhere yet (grounded)
  • c1dnn has not been integrated in any way, into the DAQ etc.  it is being run manually by hand, and will be completely shut down after this test

Simple MEDM screen I made to monitor the input/output signals:

The RTS process seems to run fine, but there is quite a bit of jitter in the CPU_METER, at the 50% level:

It's not running over the limit, but it is jumping around more than I think it should be.  Will look into that...

cpuset for cpu isolation for user-space model

The c1dnn model is running on CPU6 on c1lsc.  CPU6 was isolated from the rest of the system using cpuset.  The "cset" utility was used to create a "system" CPU set that was assigned to CPU0, and the kernel was instructed to move all running processes to that set:

controls@c1lsc:~ 2$ sudo cset set
cset:
         Name       CPUs-X    MEMs-X Tasks Subs Path
 ------------ ---------- - ------- - ----- ---- ----------
         root        0,6 y       0 y   343    0 /
controls@c1lsc:~ 0$ sudo cset set -c 0 -s system --cpu_exclusive
cset: --> created cpuset "system"
controls@c1lsc:~ 0$ sudo cset set
cset:
         Name       CPUs-X    MEMs-X Tasks Subs Path
 ------------ ---------- - ------- - ----- ---- ----------
         root        0,6 y       0 y   342    1 /
       system          0 y       0 n     0    0 /system
controls@c1lsc:~ 0$ sudo cset proc --move -f root -t system -k
cset: moving all tasks from root to /system
cset: moving 292 userspace tasks to /system
cset: moving 0 kernel threads to: /system
cset: --> not moving 50 threads (not unbound, use --force)
[==================================================]%
cset: done
controls@c1lsc:~ 0$ sudo cset set
cset:
         Name       CPUs-X    MEMs-X Tasks Subs Path
 ------------ ---------- - ------- - ----- ---- ----------
         root        0,6 y       0 y    50    1 /
       system          0 y       0 n   292    0 /system
controls@c1lsc:~ 0$ sudo cset proc --move -f root -t system -k --force
cset: moving all tasks from root to /system
cset: moving 50 kernel threads to: /system
[==================================================]%
cset: **> 29 tasks are not movable, impossible to move
cset: done
controls@c1lsc:~ 0$ sudo cset set
cset:
         Name       CPUs-X    MEMs-X Tasks Subs Path
 ------------ ---------- - ------- - ----- ---- ----------
         root        0,6 y       0 y    29    1 /
       system          0 y       0 n   313    0 /system
controls@c1lsc:~ 0$

I then created a set for the RTS process ("rts-c1dnn") on CPU6, and executed the c1dnn model in that set:

controls@c1lsc:~ 0$ sudo cset set -c 6 -s rts-c1dnn --cpu_exclusive
cset: --> created cpuset "rts-c1dnn"
controls@c1lsc:~ 0$ sudo cset set
cset:
         Name       CPUs-X    MEMs-X Tasks Subs Path
 ------------ ---------- - ------- - ----- ---- ----------
         root        0,6 y       0 y    24    2 /
    rts-c1dnn          6 y       0 n     0    0 /rts-c1dnn
       system          0 y       0 n   340    0 /system
controls@c1lsc:~ 0$ sudo cset proc -s rts-c1dnn --exec /opt/rtcds/caltech/c1/target/c1dnn/bin/c1dnn -- -m c1dnn
cset: --> last message, executed args into cpuset "/rts-c1dnn", new pid is: 27572
sysname = c1dnn
....

When done I just hit Ctrl-C.

I left the cpusets as they are, with all system processes in the "system" set.  This should not pose any problems since it's the identical configuration as would be if a normal kernel-level model was running in CPU6.

The c1dnn process and it's EPICS sequencer were shutdown after this test.

  8816   Tue Jul 9 23:27:17 2013 KojiSummaryLSCMICH: ITMX/Y <=> PRM/BS

The MICH actuation with PRM/BS was investigated again.

(ITMX -1 / ITMY +1) is equivalent to (PRM -0.267 and BS +0.50).


- PRMIsb was locked with REFL33I&AS55Q.

- Using the locking module in the LSC model, actuate ITMX (-1) and ITMY (+1) at 580.1Hz. Note that the notch filters in the MICH/PRCL servos were on.

- Look at the peak in the AS55Q spectrum. Tune the BS element in the output matrix of the lock-in to minimize the peak height.
=> The peak was minimized at BS = -0.50.

- Look at the peak in the REFL33I spectrum. Tune the PRM element in the output matrix of the lock-in to minimize the peak height.
=> The peak was minimized at PRM = +0.267

- These measurement leads to the conclusion mentioned above.

  8490   Thu Apr 25 04:10:09 2013 JenneUpdateLockingMICH_CTRL drifting away??

Koji is elogging separately of his exploration of different configurations.  The lock stretch that I'm looking at here uses the same parameters as Koji had for PRMI sb lock, using AS55Q for MICH and REFL33I for PRCL, with MICH gain of -0.8 and PRCL gain of 0.05 .

All of these plots are the same few second lock stretch, with different zooming.  Jamie's super-sweet getdata python script only accepts integers for the start time and duration parameters, so lots of this zooming happened by hand, but I tried to always keep the time axis aligned within each screenshot.  Sometimes the plot axis labels say differently, but they're lying to you.

Plot 1:  gps start time is 1050915916, duration = 6 seconds.  Overall view of the lock stretch.

1050915916-6.png

Plot 2:  gps start time is 1050915921, duration = 1 second.  We're looking at the lockloss that happens at the left side of the plots.

1050915921-1.png

Plot 3:  zoomed in (along the time-axis) version of plot 2, so much shorter time duration.  Some zooming on y-axes.

1050915921-zoom.png

Plot 4:  zoomed in (along y-axes) version of plot 2.

1050915921-1-zoom.png

It seems to me from these plots that maybe MICH CTRL is drifting away?  It seems like we lose the MICH lock, and that destroys the whole thing. 

Koji made some comments to me earlier, regarding his work this evening, that the MICH signal quality is poor in general, and that we should calculate/think about changing our schnupp asymmetry. 

  11499   Wed Aug 12 16:39:46 2015 IgnacioUpdateIOOMISO WIener (T240-X and T240-Y) FF of MCL

Last night I performed some MISO FF on MCL using the T240-X and T240-Y as witnesses. Here are the results:

Filter:

T240-X

T240-Y

 

Training data + Predicted FIR and IIR subtraction:

Online subtraction results:

MCL
YARM

Subtraction performace:

Attachment 1: stsx.png
stsx.png
Attachment 2: stsy.png
stsy.png
Attachment 3: performance.png
performance.png
Attachment 4: sub.png
sub.png
Attachment 5: mcliir.png
mcliir.png
Attachment 6: yarmiir.png
yarmiir.png
  11547   Sun Aug 30 23:47:02 2015 IgnacioUpdateIOOMISO Wiener Filtering of MCL

I decided to give MISO Wiener filtering a try again. This time around I managed to get working filters. The overall performance of these MISO filters is much better than the SISO I constructed on elog:11541 .

The procedure I used to develope the SISO filters did not work well for the construction of these MISO filters. I found a way, even more systematic than what I had before to work around Vectfit's annoyances and get the filters in working condition. I'll explain it in another eLOG post.

Anyways, here are the MISO filters for MCL using the T240-X and T240-Y as witnesses:

 Now the theoretical offline prediction:

 

 

The online subtractions for MCL, YARM and XARM. I show the SISO subtraction for reference.

 And the subtraction performance:

  11549   Mon Aug 31 09:36:05 2015 IgnacioUpdateIOOMISO Wiener Filtering of MCL

MISO Wiener filters for MCL kept the mode cleaner locked for a good 8+ hours.

  15363   Tue Jun 2 14:05:24 2020 HangUpdateBHDMM telescope actuation range requirments

We computed the required actuation range for the telescope design in elog:15357. The result is summarized in the table below. Here we assume we misalign an IFO mirror by 1 urad, and then compute how many urad do we need to move the (AS1, AS4) or (LO1, LO2) mirrors to simultaneously correct for the two gouy phases. 

Actuation requirement in urad per urad misalignment
[urad/urad] ITMX ITMY ETMX ETMY BS PRM PR2 PR3 SR3 SRM
AS1 1.9 2.1 -5.0 -5.5 0.5 0.5 -0.3 0.2 0.1 0.6
AS4 2.9 2.0 -8.8 -5.5 -5.9 -0.7 1.3 -0.7 -0.5 0.7
LO1 -4.0 -3.9 11.0 10.4 1.9 -0.4 -0.2 0.1 0.0 -1.1
LO2 -5.0 -3.7 15.1 10.4 8.7 0.8 1.9 1.1 0.7 -1.3

The most demanding ifo mirrors are the ETMs and the BS, for every 1 urad misalignment the telescope needs to move 10-15 urad to correct for that. However, it is unlikely for those mirrors to move more 100 nrad for a locked ifo with ASC engaged. Thus a few urad actuation should be sufficient. For the recycling mirrors, every 1 urad misalignment also requires ~ 1 urad actuation. 

As a result, if we could afford 10 urad actuation range for each telescope suspension, then the gouy phase separations we have should be fine. 

================================================================

Edits:

We looked at the oplev spectra from gps 1274418500 for 512 sec. This should be a period when the ifo was locked in the PRFPMI state according to elog:15348. We just focused on the yaw data for now. Please see the attached plots. The solid traces are for the ASD, and the dotted ones are the cumulative rms. The total rms for each mirror is also shown in the legend. 

I am now confused... The ITMs looked somewhat reasonable in that at least the < 1 Hz motion was suppressed. The total rms is ~ 0.1 urad, which was what I would expect naively (~ x100 times worse than aLIGO). 

There seems to be no low-freq suppression on the ETMs though... Is there no arm ASC at the moment???

Attachment 1: TM_OL_spec_1274418500_512.pdf
TM_OL_spec_1274418500_512.pdf
Attachment 2: CORNER_OL_spec_1274418500_512.pdf
CORNER_OL_spec_1274418500_512.pdf
  15386   Tue Jun 9 14:55:43 2020 JonUpdateBHDMM telescope actuation range requirments

I don't think we ever discussed why the angular RMS of the ETMs is so much higher than the ITMs. Maybe that's a separate matter because, even assuming the worst case, the actuation range requirement is

(0.82 μrad RMS) x (15 μrad/μrad) x (10 safety factor) = 0.12 mrad

which is still only order 1% of the pitch/yaw pointing range of the Small Optic Suspensions, according to P1600178 (sec. IV. A). Can we check this requirement off the list?

Quote:

We computed the required actuation range for the telescope design in elog:15357. The result is summarized in the table below. Here we assume we misalign an IFO mirror by 1 urad, and then compute how many urad do we need to move the (AS1, AS4) or (LO1, LO2) mirrors to simultaneously correct for the two gouy phases. 

Actuation requirement in urad per urad misalignment
[urad/urad] ITMX ITMY ETMX ETMY BS PRM PR2 PR3 SR3 SRM
AS1 1.9 2.1 -5.0 -5.5 0.5 0.5 -0.3 0.2 0.1 0.6
AS4 2.9 2.0 -8.8 -5.5 -5.9 -0.7 1.3 -0.7 -0.5 0.7
LO1 -4.0 -3.9 11.0 10.4 1.9 -0.4 -0.2 0.1 0.0 -1.1
LO2 -5.0 -3.7 15.1 10.4 8.7 0.8 1.9 1.1 0.7 -1.3

The most demanding ifo mirrors are the ETMs and the BS, for every 1 urad misalignment the telescope needs to move 10-15 urad to correct for that. However, it is unlikely for those mirrors to move more 100 nrad for a locked ifo with ASC engaged. Thus a few urad actuation should be sufficient. For the recycling mirrors, every 1 urad misalignment also requires ~ 1 urad actuation. 

As a result, if we could afford 10 urad actuation range for each telescope suspension, then the gouy phase separations we have should be fine. 

================================================================

Edits:

We looked at the oplev spectra from gps 1274418500 for 512 sec. This should be a period when the ifo was locked in the PRFPMI state according to elog:15348. We just focused on the yaw data for now. Please see the attached plots. The solid traces are for the ASD, and the dotted ones are the cumulative rms. The total rms for each mirror is also shown in the legend. 

I am now confused... The ITMs looked somewhat reasonable in that at least the < 1 Hz motion was suppressed. The total rms is ~ 0.1 urad, which was what I would expect naively (~ x100 times worse than aLIGO). 

There seems to be no low-freq suppression on the ETMs though... Is there no arm ASC at the moment???

 

  8084   Thu Feb 14 10:42:41 2013 JamieSummaryAlignmentMMT, curved TTs does not explain beam ellipticity at Faraday

After using alamode to calculate the round-trip mode of the beam at the Faraday exit after retro-reflection form the PRM, I'm not able to blame the MMT and TT curvature for the beam ellipticity.

I assume an input waist at the mode cleaner of [0.00159, 0.00151] (in [T, S]).  Propagating this through the MMT to PRM, then retro-reflecting back with flat TTs I get

w_t/w_s = 0.9955,  e = 0.0045

If I give the TTs a -600 m curvature, I get:

w_t/w_s = 1.0419,  e = 0.0402

That's just a 4% ellipticity, which is certainly less than we see.  I would have to crank up the TT curvature to -100m or so to see an ellipticity of 20%.  We're seeing something that looks bigger than 50% to me.

Below are beam size through MMT + PRM retro-reflection, TT RoC = -600m:

 

T.pdfS.pdf

  3578   Wed Sep 15 16:12:35 2010 koji, steveUpdateMOPAMOPA Controller is taken out of the PSL rack

We removed the Lightwave MOPA Controller, PA#102, NPRO206 power supply to make room for the IOO chassy at 1X1 (south) rack.

The umbilical cord was a real pain to take out. It is shading its plastic cover. The unused Minco was disconnected and removed.

The ref. cavity ion pump controller- power supply was temporarily taken out also.

Attachment 1: P1060843.JPG
P1060843.JPG
  663   Sun Jul 13 17:19:29 2008 ranaSummaryPSLMOPA SLOWM Calibration
John, Rana

We first unlocked the FSS and ramped the SLOW actuator. With the PMC locked we observed the PMC PZT voltage
as a function of SLOWM (SLOW loop actuator voltage). We believed this to be ~1-5 GHz / V. Since this is
not so precise we then ran a slow (2 min. period) triangle wave into the slow actuator and looked at the
ref cav transmission peaks to calibrate it.

Plot is attached>

We assume that the reference cavity length = 203.2 mm then the FSR = 737.7 MHz. So looking at the plot
and using our eye to measure the SLOWM calibration is 1054 +/- 30 MHz/V. This is probably dominated by
our eye method.

Note: we tried to get the length from T010159-00-R (Michele, Weinstein, Dugolini). In that doc,
the length used is 203.3 mm whereas its 203.2 mm in the PSL FDD (?). The calculation of the FSR is also
incorrect (looks like they used c = 299460900 instead of 299792458 m/s). We took the length from the PSL FDD
(T990025-00-D) but not the FSR, since they also did not find the right value of 'c'. I guess that the speed
of light just ain't what it used to be.
Attachment 1: SLOWDCcalibration.png
SLOWDCcalibration.png
  1573   Mon May 11 11:49:20 2009 steveUpdatePSLMOPA cooling water lines are backwards

Quote:
This is 8 days of 10-minute trend.

DTEC is just the feedback control signal required to keep the NPRO's pump diode at a constant temperature.
Its not the amplifier or the actual NPRO crystal's temperature readout.

There is no TEC for the amplifier. It looks like to me that by opening up the flow to the NPRO some more
we have reduced the flow to the amplifier (which is the one that needs it) and created these temperature
fluctuations.

What we need to do is choke down the needle valve and ream out the NPRO block.




I have measured the "input" line temp at the MOPA box 10 C and the "out" line 8 C

This must be corrected.

However look at the 80 days plot of operation where the head temp variation is nothing new
Attachment 1: htempvar80d.jpg
htempvar80d.jpg
  1281   Fri Feb 6 16:20:52 2009 YoichiUpdatePSLMOPA current slider fixed

I fixed the broken slider to change the current of the PA.

The problem was that the EPICS database assigned a wrong channel of the DAC to the slider.

I found that the PA current adjustment signal lines are connected to the CH3 &CH4 of VMIC4116 #1. However in the database file (/cvs/cds/caltech/target/c1psl/psl.db), the slider channel (C1:PSL-126MOPA_DCAMP) was assigned to CH2. I fixed the database file and rebooted c1psl. Then the PA current started to follow the slider value.

I moved the slider back and forth by +/-0.3V while the ISS loop was on. I observed that the amount of the low frequency fluctuation of the MOPA power changed with the slider position. At some current levels, the ISS instability problem went away.

Kakeru is now taking open-loop TFs and current shunt responses at different slider settings.

  2133   Thu Oct 22 15:44:16 2009 ZachUpdateWIKI-40M UpdateMOPA diagram

 I have updated the PSL Diagram wiki page to include MOPA. As with the PSL diagram, clicking the photo on the main page takes you to a larger image. The inventory is pretty meager as I didn't have time to sit and read labels (if indeed there are any). I will look through the documentation at the 40m to see if there is a record of what is there. Again, if you know something, please amend the list!!

http://lhocds.ligo-wa.caltech.edu:8000/40m/PSL_Table_Diagram

  1254   Wed Jan 28 12:42:51 2009 YoichiUpdatePSLMOPA dying
Yoichi, Jenne, Peter

As most of you know, the MOPA output power has been declining rapidly since Jan 21. (See the attachment 1)
There was also an increase in the NPRO power observed in LMON, which is an internal power monitor of the NPRO.
Similar trend can be seen in 126MON, which picks up some scattered light from the NPRO but there may be some contributions from the PA output.

The drop in the AMPMON, LMON and CURMON (NPRO current) from the middle of Jan 26 to the end of Jan27 was caused by me.
I tried to decrease the NPRO current to put the NPRO power back to the level when the MOPA output was higher. But it did not bring back the MOPA power.
So I put back the current after an hour. This caused the sharp power drop on Jan26.
By mistake, I did not fully recover the current at that time and left it like that for a day. This accounts for the long power drop period continued until Jan27.

Shortly after I tweaked the current, the MOPA output power started to fluctuate a lot. This drives the ISS crazy.
To see if this was caused by the NPRO or power amplifier,
we decided to fix the 126MON to monitor the real NPRO power.
We opened the MOPA box and installed a mirror to direct a picked off NPRO beam to the outside of the box through an unused hole.
We set up a lens and a PD outside of the MOPA box to receive this beam. The output from the PD is connected to the 126MON cable.
So 126MON is now serving as the real monitor of the NPRO power. It has not yet been calibrated.

The second attachment shows a short time series of the MOPA power and NPRO power. When the beam is blocked, the 126MON goes to -22.
So the RIN of the NPRO is less than 1%, whereas the MOPA power fluctuates about 5%. There is also no clear correlation between the power fluctuation of the MOPA and the NPRO. So probably the MOPA power fluctuation is not caused by NPRO.

At this moment, all the feedback signals (current shunt, slow and fast actuators) are physically disconnected from MOPA box so that we can see the behavior of MOPA itself.
Attachment 1: Recent10Days.png
Recent10Days.png
Attachment 2: 126_MOPA.png
126_MOPA.png
  3132   Tue Jun 29 10:20:58 2010 ranaUpdateMOPAMOPA is NOT dead

Not dead. It just had a HT fault. You can tell by reading the front panel. Cycling the power usually fixes this.

  3137   Tue Jun 29 16:44:12 2010 Jenne, ranaUpdateMOPAMOPA is NOT dead, was just asleep

Quote:

Not dead. It just had a HT fault. You can tell by reading the front panel. Cycling the power usually fixes this.

MOPA is back onliine.  Rana found that the fuse in the AC power connector's fuse had blown.  This was evident by smelling all of the inputs and outputs of the MOPA controller. The power cord we were using for this was only rated for 10A and therefore was a safety hazard. The fuse should be rated to blow before the power cord catches on fire. The power cord end was slightly melted. I don't know why it hadn't failed in the last 12 years, but I guess the MOPA was drawing a lot of extra current for the DTEC or something due to the high temperature of the head.

We got some new fuses from Todd @ Downs. 

The ones we got however were fast-blow, and that's what we want  The fuses are 10A, 250V.  The fuses are ~.08 inches long, 0.2 inches in diameter. 

  3130   Tue Jun 29 08:41:06 2010 steveUpdateMOPAMOPA is dead

I found the laser dead this morning.

The crane people are here to unjam it.

Laser hazard mode is lifted and LASER SAFE MODE is in place. No safety glasses but CRANE HAZARD is still active.

Stay out of the 40m lab !

 

 

Attachment 1: laserisdead.jpg
laserisdead.jpg
  1042   Mon Oct 13 11:32:50 2008 YoixhiUpdatePSLMOPA is in trouble now
Steve, Alberto, Yoichi

A quick update.
The MOPA output went down to zero on Sunday early morning (00:28 AM).
We found that the NPRO beam is mis-aligned on the power monitoring PD (126MON).
We don't know yet if it is also mis-aligned to the power amplifier (PA) because the mechanical shutter is not working (always closed).
Most likely the beam is not aligned to the PA.
A mystery is that although the beam is terribly (more than a half inch) missing the monitor PD, the beam still goes through two faradays.
Another mystery is that the NPRO output power is now increased to 600mW.

The power drop was a very fast phenomenon (less than 1/16 sec).
We are trying to figure out what happened.
The first step is to fix the mechanical shutter. We have a spare in our hand.
Attachment 1: powerdrop.png
powerdrop.png
ELOG V3.1.3-