ID Date Author Type Category Subjectup
  14454   Thu Feb 14 21:29:24 2019 gautamSummaryLoss MeasurementInferred Y arm loss

Summary:

From the measurements I have, the Y arm loss is estimated to be 58 +/- 12 ppm. The quoted values are the median (50th percentile) and the distance to the 25th and 75th quantiles. This is significantly worse than the ~25 ppm number Johannes had determined. The data quality is questionable, so I would want to get some better data and run it through this machinery and see what number that yields. I'll try and systematically fix the ASS tomorrow and give it another shot.

Model and analysis framework:

Johannes and I have cleaned up the equations used for this calculation - while we may make more edits, the v1 of the document lives here. The crux of it is that we would like to measure the quantity \kappa = \frac{P_L}{P_M}, where P_{L(M)} is the power reflected from the resonant cavity (just the ITM). This quantity can then be used to back out the round-trip loss in the resonant cavity, with further model parameters which are:

  1. ITM and ETM power transmissivities
  2. Modulation depths and mode-matching efficiency into the cavity
  3. The statistical uncertainty on the measurement of the quantity \kappa, call it \sigma_{\kappa}

If we ignore the 3rd for a start, we can calculate the "expected" value of \kappa as a function of the round-trip loss, for some assumed uncertainties on the above-mentioned model parameters. This is shown in the top plot in Attachment #1, and while this was generated using emcee, is consistent with the first order uncertainty propagation based result I posted in my previous elog on this subject. The actual samples of the model parameters used to generate these curves are shown in the bottom. What this is telling us is that even if we have no measurement uncertainty on \kappa, the systematic uncertainties are of the order of 5 ppm, for the assumed variation in model parameters.

The same machinery can be run backwards - assuming we have multiple measurements of \kappa, we then also have a sample variance, \sigma_{\kappa}. The uncertainty on the sample variance estimator is also known, and serves to quantify the prior distribution on the parameter \sigma_{\kappa} for our Monte-Carlo sampling. The parameter \sigma_{\kappa} itself is required to quantify the likelihood of a given set of model parameters, given our measurement. For the measurements I did this week, my best estimate of \kappa \pm \sigma_{\kappa} = 0.995 \pm 0.005. Plugging this in, and assuming uncorrelated gaussian uncertainties on the model parameters, I can back out the posterior distributions.

For convenience, I separate the parameters into two groups - (i) All the model parameters excluding the RT loss, and (ii) the RT loss. Attachment #2 and Attachment #3 show the priors (orange) and posteriors (black) of these quantities. 

Interpretations:

  1. This particular technique only gives us information about the RT loss - much less so about the other model parameters. This can be seen by the fact that the posteriors for the loss is significantly different from the prior for the loss, but not for the other parameters. Potentially, the power of the technique is improved if we throw other measurements at it, like ringdowns.
  2. If we want to reach the 5 ppm uncertainty target, we need to do better both on the measurement of the DC reflection signals, and also narrow down the uncertainties on the other model parameters.

Some assumptions:

So that the experts on MC analysis can correct me wheere I'm wrong.

  1. The prior distributions are truncated independent Gaussians - truncated to avoid sampling from unphysical regions (e.g. negative ITM transmission). I've not enforced the truncation analytically - i.e. I just assume a -infinity probability to samples drawn from the unphysical parts, but to be completely sure, the actual cavity equations enforce physicality independently (i.e. the MC generates a set of parameters which is input to another function, which checks for the feasibility before making an evaluation). One could argue that the priors on some of these should be different - e.g. uniform PDF for loss between some bounds? Jeffrey's prior for \sigma_{\kappa}?
  2. How reasonable is it to assume the model parameter uncertainties are uncorrelated? For exaple, \eta, \beta_1, \beta_2 are all determined from the ALS-controlled cavity scan
Attachment 1: modelPerturb.pdf
modelPerturb.pdf
Attachment 2: posterior_modelParams.pdf
posterior_modelParams.pdf
Attachment 3: posterior_Loss.pdf
posterior_Loss.pdf
  15163   Tue Jan 28 14:33:24 2020 gautamUpdatePSLInferred free-running frequency noise

To conclude my PMC noise investigations: Attachment #1 shows the PMC noise inferred from the calibrations earlier in this thread and the fitted OLTF for the PMC loop. Attachment #2 compares the frequency noise (inferred from the error point of the PMC servo) when the IMC is locked / unlocked. I don't know what to make of the fact that the PMC suggests improvement from ~20 Hz onwards already - does this mean that the NPRO noise model is wrong by 1 order of magnitude at 30 Hz?

  • The IMC was locked for the measurement shown in Attachment #1.
  • The in-loop spectra of the error (at the I/F output of the mixer) and control (at TP3) signals were measured with the SR785.
  • The control signal voltage monitors don't seem to work - neither the front panel LEMO nor the signals hooked up to the CDS system show me sensible shapes for the spectra between 1-3 Hz.
  • To convert in loop to free-running, I multiplied the measured error (control) signal spectra by \left | 1-L \right | (\left | \frac{L}{1-L} \right |), where L is the OLTF. THe control signal was pre-processed by multiplying by a pole at 11.3 Hz, corresponding to the LPF formed by the 63.3 kohm series resistor and the 225 nF PZT capacitance.
  • The "NPRO noise model" curve is 10^4/f Hz/rtHz.

While I initially thought the 1/f^2 rise below ~100 Hz is attributable to the IMC cavity length fluctuations, I found that this profile is present even in the measurement with the PSL shutter closed. I am not embarking on a detailed PMC noise budgeting project for now. Note however that we are not shot noise limited anywhere in this measurement band.

  • The measured TF (dots in Attachment #5) was fit with LISO (solid lines in Attachment #5) to allow inferring the out-of-loop servo noise by monitoring the in-loop noise (that plot to follow).
Attachment 1: inLoopNoise_IMClocked.pdf
inLoopNoise_IMClocked.pdf
Attachment 2: freqNoiseComparison.pdf
freqNoiseComparison.pdf
  16323   Mon Sep 13 17:05:04 2021 TegaSummaryPEMInfrasensing temperature sensor modbus configuration

Anchal mentioned it would be good to put more details about how I arrived at the values needed to configure the modbus drive for the temperature sensor, since this information is not in the manual and is hard to find on the internet, so here is a breakdown.

So the generic format is:

drvAsynIPPortConfigure("<TCP_PORT_NAME>","<UNIT_IP_ADDRESS>:502",0,0,1)
modbusInterposeConfig("<TCP_PORT_NAME>",0,5000,0)
drvModbusAsynConfigure("<PORT_NAME>","<TCP_PORT_NAME>",<slaveAddress>,<modbusFunction>,<modbusStartAddress>,<modbusLength>,<dataType>,<pollMsec>,<plcType>)

which in our case become:

drvAsynIPPortConfigure("c1pemxendtemp","192.168.113.240:502",0,0,1)
modbusInterposeConfig("c1pemxendtemp",0,5000,0)
drvModbusAsynConfigure("C1PEMXENDTEMP","c1pemxendtemp",0,4,199,2,0,1000,"ServerCheck")

As can be seen, the parameters of the first two functions "drvAsynIPPortConfigure" and "modbusInterposeConfig" are straight forward, so we restrict our discussion to the case of third function "drvModbusAsynConfigure". Well, after hours of trolling the internet, I was able to piece together a coherent picture of what needs doing and I have summarised them in the table below.

 

drvModbusAsynConfigure

Once the asyn IP or serial port driver has been created, and the modbusInterpose driver has been configured, a modbus port driver is created with the following command:

drvModbusAsynConfigure(portName,                # used by channel definitions in .db file to reference this unit)
                       tcpPortName,             # reference to portName created with drvAsynIPPortConfigure command
                       slaveAddress,            # 
                       modbusFunction,          # 
                       modbusStartAddress,      # 
                       modbusLength,            # length in dataType units
                       dataType,                # 
                       pollMsec,                # how frequently to request a value in [ms]
                       plcType);                #

drvModbusAsynConfigure command
Parameter Data type Description
portName string Name of the modbus port to be created.
 
tcpPortName string Name of the asyn IP or serial port previously created.

tcpPortName = { 192.168.113.240:502192.168.113.241:502192.168.113.242:502 }
 
slaveAddress int The address of the Modbus slave. This must match the configuration of the Modbus slave (PLC) for RTU and ASCII. For TCP the slave address is used for the "unit identifier", the last field in the MBAP header. The "unit identifier" is ignored by most PLCs, but may be required by some.

ServersCheck API ignores this value, as confirmed with pymodbus query, so set to default value: 
slaveAddress = 0
 
modbusFunction int

modbus supports the following 8 Modbus function codes:

Modbus Function Codes
Access Function description Function code
Bit access Read Coils 1
Bit access Read Discrete Inputs 2
Bit access Write Single Coil 5
Bit access Write Multiple Coils 15
16-bit word access Read Input Registers 4
16-bit word access Read Holding Registers 3
16-bit word access Write Single Register 6
16-bit word access Write Multiple Registers 16
modbusStartAddress int Start address for the Modbus data segment to be accessed.
(0-65535 decimal, 0-0177777 octal).

Modbus addresses are specified by a 16-bit integer address. The location of inputs and outputs within the 16-bit address space is not defined by the Modbus protocol, it is vendor-specific. Note that 16-bit Modbus addresses are commonly specified with an offset of 400001 (or 300001). This offset is not used by the modbus driver, it uses only the 16-bit address, not the offset.

For ServersCheck, the offset is "30001", so that

modbusStartAddress = 30200 - 30001 = 199

modbusLength int The length of the Modbus data segment to be accessed.
This is specified in bits for Modbus functions 1, 2, 5 and 15.
It is specified in 16-bit words for Modbus functions 3, 4, 6 and 16.
Length limit is 2000 for functions 1 and 2, 1968 for functions 5 and 15,
125 for functions 3 and 4, and 123 for functions 6 and 16.

ServersCheck uses two's complement 32-bits word (with big-endian byte order & little-endian word order) format to store floating-point data, as confirmed with pymodbus query, so that:

modbusLength = 2
 
modbusDataType int The modbusDataType is used to tell the driver the format of the Modbus data. The driver uses this information to convert the number between EPICS and Modbus. Data is transferred to and from EPICS as epicsUInt32, epicsInt32, and epicsFloat64 numbers.

Modbus data type:
0 = binary, twos-complement format
1 = binary, sign and magnitude format
2 = BCD, unsigned
3 = BCD, signed

Some Modbus devices (including ServersCheck) use floating point numbers, typically by storing a 32-bit float in two consecutive 16-bit registers. This is not supported by the Modbus specification, which only supports 16-bit registers and single-bit data. The modbus driver does not directly support reading such values, because the word order and floating point format is not specified.

Note that if it is desired to transmit BCD numbers untranslated to EPICS over the asynInt32 interface, then data type 0 should be used, because no translation is done in this case. 

For ServersCheck, we wish to transmit the untranslated data, so:

modbusDataType = 0
 
pollMsec int Polling delay time in msec for the polling thread for read functions.
For write functions, a non-zero value means that the Modbus data should
be read once when the port driver is first created.

ServersCheck recommends setting sensor polling interval between 1-5 seconds, so we can try:

pollMsec = 1000
 
plcType string Type of PLC (e.g. Koyo, Modicon, etc.).
This parameter is currently used only to print information in asynReport.
In the future it could be used to modify the driver behavior for a specific PLC.

plcType = "ServersCheck"
 

 

Useful links

https://nodus.ligo.caltech.edu:8081/40m/16214

https://nodus.ligo.caltech.edu:8081/40m/16269

https://nodus.ligo.caltech.edu:8081/40m/16270

https://nodus.ligo.caltech.edu:8081/40m/16274

 

http://manuals.serverscheck.com/InfraSensing_Sensors_Platform.pdf

http://manuals.serverscheck.com/InfraSensing_Modbus_manualv5.pdf

https://community.serverscheck.com/discussion/comment/7419#Comment_7419

 

https://wiki-40m.ligo.caltech.edu/CDS/SlowControls

https://www.slac.stanford.edu/grp/ssrl/spear/epics/site/modbus/modbusDoc.html#Creating_a_modbus_port_driver

 

https://github.com/riptideio/pymodbus

 

https://en.wikipedia.org/wiki/Modbus

https://deltamotion.com/support/webhelp/rmctools/Communications/Ethernet/Supported_Protocols/Ethernet_Modbus_TCP.htm

  16761   Thu Apr 7 11:47:22 2022 YehonathanUpdateBHDInitial BHD modeling: AS - LO mode matching

I begin modeling the initial BHD setup using Finesse. I started with copying C1_w_BHD.kat from the 40m/bhd repo and making changes to reflect the current BHD setup:

1. OMCs were removed.

2. Only 1 PD per BHD port was left.

3. Transmission of PR2 was changed to 2.2%. The PRG was calculated to be ~15.5.

4. Actual RoCs of new optics were dialed in (Yesterday me and Paco went into the cleanroom to measure the RoCs and they seem to match the datasheets).

Here's a table comparing the old (design?) RoCs with the new RoCs:

  New RoC Old RoC
LO1 5m 6m
LO2 inf inf
LO3 500mm 750mm
LO4 150mm -450mm
AS1 2m 2.8m
AS2 inf inf
AS3 200mm -2m
AS4 750mm 600mm
PR2 2000m -700m
PR3 1000m 1000m
SR3 1000m -700m

 

The changes looked quite alarming, especially for LO4 and AS3, so I wrote a script to calculate the mode matching between the LO and AS beams called AS_LO_ModeMatching.ipynb and pushed it into the repo. In the notebook a bright AS beam is created by creating a small asymmetry between the arms of ~ 0.003 degrees (~10pm). Amplitude detectors were put at the input ports of the BHD BS to calculate the fields in the AS and LO beams. In particular TEM00, TEM02 and TEM20 were measured for each beam.

The calculation shows that with the old RoCs the mode matching between the LO and AS beams is 99% while for the new RoCs it is ~ 50%.

  16766   Thu Apr 7 21:15:04 2022 YehonathanUpdateBHDInitial BHD modeling: AS - LO mode matching

Ok, it turns out these optics were purchased on purpose, as this elog shows. Jon considered building a mode-matching telescope with stock optics as an initial step before purchasing the custom optics (referred to as "design" optics in my elog).

I dialed in the new distances between the optics into the .kat file as described in this elog and pushed the changes to the repo. With the new distances, I got mode-matching of 87% for the full IFO and 89% for FPMI so there's probably no need to worry as the mode-matching with these optics was already designed.

Quote:

I begin modeling the initial BHD setup using Finesse. I started with copying C1_w_BHD.kat from the 40m/bhd repo and making changes to reflect the current BHD setup:

1. OMCs were removed.

2. Only 1 PD per BHD port was left.

3. Transmission of PR2 was changed to 2.2%. The PRG was calculated to be ~15.5.

4. Actual RoCs of new optics were dialed in (Yesterday me and Paco went into the cleanroom to measure the RoCs and they seem to match the datasheets).

Here's a table comparing the old (design?) RoCs with the new RoCs:

  New RoC Old RoC
LO1 5m 6m
LO2 inf inf
LO3 500mm 750mm
LO4 150mm -450mm
AS1 2m 2.8m
AS2 inf inf
AS3 200mm -2m
AS4 750mm 600mm
PR2 2000m -700m
PR3 1000m 1000m
SR3 1000m -700m

 

The changes looked quite alarming, especially for LO4 and AS3, so I wrote a script to calculate the mode matching between the LO and AS beams called AS_LO_ModeMatching.ipynb and pushed it into the repo. In the notebook a bright AS beam is created by creating a small asymmetry between the arms of ~ 0.003 degrees (~10pm). Amplitude detectors were put at the input ports of the BHD BS to calculate the fields in the AS and LO beams. In particular TEM00, TEM02 and TEM20 were measured for each beam.

The calculation shows that with the old RoCs the mode matching between the LO and AS beams is 99% while for the new RoCs it is ~ 50%.

 

  16858   Mon May 16 16:13:01 2022 YehonathanUpdateBHDInitial BHD modeling: Damped suspension model

I was finally able to set up a stable suspension model with the help of Yuta and I'm now ready to start doing some MICH noise budgeting with BHD readout. (Tip: turns out that in the zpk function in Matlab you should multiply the poles and zeros by -2*pi to match the zpk TFs in Foton)

I copied all the filters from the suspension MEDM screens into a Matlab. Those filters were concatenated with a single pendulum suspension TF with poles at [0.05e-1+1i, 0.05e-1-1i] and a gain of 4 N/kg.

I multiplied the OLTF with the real gains at the DAC/DAC/OSEMs/Coil Driver and Coils. I ignore whitening/dewhitening for now. The OLTF was calculated with no additional ad-hoc gain.

Attachment 1 shows the calculated open-loop transfer function.

Attachment 2 shows OLTF of ETMY measured last week.

Attachment 3 shows the step and impulse responses of the closed-loop system.

 

 

Attachment 1: Damped_SUS.png
Damped_SUS.png
Attachment 2: ETMY_SUSPOS_GOL.pdf
ETMY_SUSPOS_GOL.pdf ETMY_SUSPOS_GOL.pdf
Attachment 3: SUS_Response.png
SUS_Response.png
  3284   Sat Jul 24 13:13:41 2010 rana, steve, albertoUpdateGeneralInitial Crane Inspection reveals flaws: wiring and oil

The guy from KroneCrane (sp?) came today and started the crane inspection on the X End Crane. There were issues with our crane so he's going to resume on Monday. We turned off the MOPA fur the duration of the inspection.

  1. None of our cranes have oil in the gearbox and it seems that they never did since they have never been maintained. Sloppy installation job. The crane oiling guy is going to come in on Monday.
  2. They tried to test the X-End crane with 2500 lbs. (its a 1 ton crane). This tripped the thermal overload on the crane as intended with this test. Unfortunately, the thermal overload switch disabled the 'goes down' circuit instead of the 'goes up' circuit as it should. We double checked the wiring diagram to confirm our hypothesis. Seems the X-End crane was wired up incorrectly in the first place 16 years ago. We'll have to get this fixed.

The plan is that they will bring enough weight to test it at slightly over the rating (1 Ton + 10 %) and we'll retry the certification after the oiling on Monday.

  3290   Mon Jul 26 10:04:24 2010 steveUpdateGeneralInitial Crane Inspection reveals flaws: wiring, oil and imbalance

Quote:

The guy from KroneCrane (sp?) came today and started the crane inspection on the X End Crane. There were issues with our crane so he's going to resume on Monday. We turned off the MOPA fur the duration of the inspection.

  1. None of our cranes have oil in the gearbox and it seems that they never did since they have never been maintained. Sloppy installation job. The crane oiling guy is going to come in on Monday.
  2. They tried to test the X-End crane with 2500 lbs. (its a 1 ton crane). This tripped the thermal overload on the crane as intended with this test. Unfortunately, the thermal overload switch disabled the 'goes down' circuit instead of the 'goes up' circuit as it should. We double checked the wiring diagram to confirm our hypothesis. Seems the X-End crane was wired up incorrectly in the first place 16 years ago. We'll have to get this fixed.

The plan is that they will bring enough weight to test it at slightly over the rating (1 Ton + 10 %) and we'll retry the certification after the oiling on Monday.

 The south end crane has one more flaw. The wall cantilever is imbalanced: meaning it wants to rotate south ward, because its axis is off.

This effects the rope winding on the drum as it is shown on Atm2

Atm1 is showing Jay Swar of KoneCrane and the two 1250 lbs load that was used for the test. Overloading the crane at 125% is general practice at load testing.

It was good to see that the load brakes were working well at 2500 lbs. Finally we found a good service company! and thanks for Rana and Alberto

for coming in on Saturday.

Attachment 1: 2500.JPG
2500.JPG
Attachment 2: sedrum.JPG
sedrum.JPG
  360   Wed Mar 5 12:51:48 2008 JohnSummaryLSCInitial Ligo Arm finesse versus lambda
I've taken the coating recipes for the initial ligo arm cavity from Rana's web page (ligo.caltech/edu/~rana/mat/)
and plotted the finesse as a function of wavelength. There is some uncertainty over the indices of refraction but
the main conclusion remains unchanged - i.e. it appears that using other wavelengths will be difficult.
Stefan is looking at how to tune the layers of any new mirrors to make dichroic optics.
Attachment 1: FofLambdaLIGOI.jpg
FofLambdaLIGOI.jpg
  7599   Tue Oct 23 17:30:33 2012 jamie, nic, jenne, raji, manasaUpdateAlignmentInitial attempts to fix IFO alignment

We went into the vertex today to see about fixing the alignment.  The in-air access connector is in place, and we took heavy doors off of BS, ITMY, and ETMY chambers.

We started by looking at the pointing from the PZTs.  Manasa and Raji hooked up HV power supplies to the PZTs and set them to the middle of their ranges (75 V).

We installed a target on the BS cage, and new "free standing" targets made special by Steve for the SOSs on ITMY and ETMY.

Using a free-standing aperture target we looked at the beam height before PZT2.  It was a little high, so we adjusted it with PZT1.  Once that was done we looked at the beam height at PR2, and adjusted that height with PZT1.

We then tried to use the hysteresis in PR2 to adjust the beam height at ITMY.  Pushing just a little bit at the top or bottom of PR2 would repoint the beam in pitch.  This sort of works, but it's stupid.  Using this method we got the beam more or less centered vertically at ITMY.

We moved on to ETMY with the idea that we would again use the hysteresis in PR3 to get the vertical pointing to the ETM correct.  This was a good demonstration of just how stupid the tip-tilts really are.  Just touching slightly at the top or bottom or PR3 we could completely change the pointing at ETMY, by mili-radians (~4 cm over 40m).

At this point I cried foul.  This is not an acceptable situation.  Very little stimulation to the tip-tilts can repoint the beam inside the PR cavity.

Steve says that the TT weights, which will attach to the base of the TT mirror mounts and should help keep the mirrors vertical and not hysteretic, are being baked now and should be available tomorrow.  We therefore decided to stop what we were doing today, since we'll have to just redo it all again tomorrow once the weights are installed.

 

  12080   Fri Apr 15 23:11:49 2016 gautamUpdateGeneralInnolight 1W moved to SP table

I have moved the 1W Innolight + controller from the PSL table to the SP table for beam profiling.

  2810   Mon Apr 19 16:31:42 2010 KevinUpdatePSLInnolight 2W Laser

Koji and Kevin

We unpacked the Innolight 2W laser, took an inventory, and scanned the operations manual.

[Edit by KA]

The scanned PDFs are placed on the following wiki page

http://lhocds.ligo-wa.caltech.edu:8000/40m/Upgrade_09/PSL

We will measure the P-I curve, the mode profile, frequency actuator responses, and so on.

  2822   Tue Apr 20 20:15:37 2010 KevinUpdatePSLInnolight 2W Output Power vs Injection Current

Koji and Kevin measured the output power vs injection current for the Innolight 2W laser.

The threshold current is 0.75 A.

 

The following data was taken with the laser crystal temperature at 25.04ºC (dial setting: 0.12).

Injection Current (A) Dial Setting Output Power (mW)
0.000 0.0 1.2
0.744 3.66 1.1
0.753 3.72 4.6
0.851 4.22 102
0.954 4.74 219
1.051 5.22 355
1.151 5.71 512
1.249 6.18 692
1.350 6.64 901
1.451 7.08 1118
1.556 7.52 1352
1.654 7.92 1546
1.761 8.32 1720
1.853 8.67 1855
1.959 9.05 1989
2.098 9.50 2146

 

Attachment 1: PvsI_2W.jpg
PvsI_2W.jpg
  2828   Wed Apr 21 21:56:27 2010 KevinUpdatePSLInnolight 2W Vertical Beam Profile

Koji and Kevin measured the vertical beam profile of the Innolight 2W laser at one point.

This data was taken with the laser crystal temperature at 25.04°C and the injection current at 2.092A.

The distance from the razor blade to the flat black face on the front of the laser was 13.2cm.

The data was fit to the function y(x)=a*erf(sqrt(x)*(x-x0)/w)+b with the following results.

Reduced chi squared = 14.07

x0 = (1.964 +- 0.002) mm

w  = (0.216 +- 0.004) mm

a  = (3.39  +- 0.03) V

b  = (3.46  +- 0.03) V

Attachment 1: bp2.jpg
bp2.jpg
Attachment 2: bp2.dat
razor height (mm)   Voltage (V)
2.75    6.89
2.50    6.90
2.30    6.89
2.25    6.89
2.20    6.75
2.15    6.47
2.13    6.20
2.10    6.05
2.07    5.88
... 17 more lines ...
  2829   Wed Apr 21 22:11:48 2010 ranaUpdatePSLInnolight 2W Vertical Beam Profile

Back in Gainesville in 1997, I learned how to do this using the chopper wheel. We had to make the assumption that the wheel's blade was moving horizontally during the time of the chop.

One advantage is that the repetitive slices reduces the random errors by a lot - you can trigger the scope and average. Another advantage is that you can download the average scope trace using USB, floppy, or ethernet instead of pencil and paper.

But, I never analyzed it in enough detail to see if there was some kind of nasty systematic error.

  2830   Wed Apr 21 23:35:37 2010 KojiUpdatePSLInnolight 2W Vertical Beam Profile

Good fit. I assumed sqrt(x) is a typo of sqrt(2).

Quote:

Koji and Kevin measured the vertical beam profile of the Innolight 2W laser at one point.

This data was taken with the laser crystal temperature at 25.04°C and the injection current at 2.092A.

The distance from the razor blade to the flat black face on the front of the laser was 13.2cm.

The data was fit to the function y(x)=a*erf(sqrt(x)*(x-x0)/w)+b with the following results.

Reduced chi squared = 14.07

x0 = (1.964 +-  0.002) mm

w = (0.216 +- 0.004) mm

a = (3.39 +- 0.03) V

b = (3.46 +- 0.03) V

 

  2834   Thu Apr 22 21:42:24 2010 AlbertoUpdatePSLInnolight 2W Vertical Beam Profile

 

 What kind of fit did you use? How are the uncertainties in the parameters obtained?

  14556   Fri Apr 19 14:06:36 2019 gautamUpdatePSLInnolight NPRO shutoff

When I got back from lunch just now, I noticed that the PMC TRANS and REFL cameras were showing no spots. I went onto the PSL table, and saw that the NPRO was in fact turned off. I turned it back on.

The laser was definitely ON when I left for lunch around 130pm, and this happend around 140pm. Anjali says no one was in the lab in between. None of the FEs are dead, suggesting there wasn't a labwide power outage, and the EX and EY NPROs were not affected. I had pulled out the diagnostics connector logged by Acromag, I'm restoring it now in the hope we can get some more info on what exactly happened if this is a recurring event. So FSS_RMTEMP isn't working from now on. Sooner we get the PSL Acromag crate together, the better...

Attachment 1: Screenshot_from_2019-04-19_14-06-11.png
Screenshot_from_2019-04-19_14-06-11.png
  14560   Fri Apr 19 20:21:52 2019 gautamUpdatePSLInnolight NPRO shutoff

Happened again at ~730pm.

The NPRO diag channels don't really tell me what happened in a causal way, but the interlock channel seems suspicious. Why is the nominal value 0.04 V? From the manual, it looks like the TGUARD is an indication of deviations between the set temperature and actual diode laser temperature. Is it normal for it to be putting out 11V?

I'm not going to turn it on again right now while I ponder which of my hands I need to chop off.

Quote:
 

I'm restoring it now in the hope we can get some more info on what exactly happened if this is a recurring event.

Attachment 1: Screenshot_from_2019-04-19_20-27-04.png
Screenshot_from_2019-04-19_20-27-04.png
  14566   Wed Apr 24 16:06:44 2019 gautamUpdatePSLInnolight NPRO shutoff

After discussing with Koji, I turned the NPRO back on again, at ~4PM local time. I first dialled the injection current down to 0A. Then powered the control unit state to "ON". Then I ramped up the power by turning the front panel dial. Lasing started at 0.5A, and I saw no abrupt swings in the power (I used PMC REFL as a monitor, there were some mode flashes which are the dips seen in the power, and the x-axis is in units of time not pump current). PMC was relocked and IMC autolocker locked the IMC almost immediately.

Now we wait and watch I guess.

Attachment 1: PMCrefl.png
PMCrefl.png
  14577   Thu Apr 25 17:31:56 2019 gautamUpdatePSLInnolight NPRO shutoff

NPRO shutoff at ~1517  local time today afternoon. Again, not many clues from the NPRO diagnostics channel, but to my eye, the D1_POW channel shows the first variation from the "steady state", followed by the other channels. This is ~0.1 sec before the other channels register some change, so I don't know how much we can trust the synchronizaiton of the EPICS data streams. I won't turn it on again for now. I did check that the little fan on the back of the NPRO controller is still rotating.

gautam 10am 4/29: I also added a longer term trend of these diagnostic channels, no clear trends suggesting a fault are visible. The y-axis units for all plots are in Volts, and the data is sampled at 16 Hz.

Quote:

Now we wait and watch I guess.

Attachment 1: EdwinShutoff20190425.png
EdwinShutoff20190425.png
Attachment 2: EdwinShutdown_zoomOut.png
EdwinShutdown_zoomOut.png
  12109   Thu May 5 21:28:44 2016 gautamUpdateendtable upgradeInnolight PZT capacitance

I suggested in an earlier elog that after the repair of the NPRO, the PZT capacitance may have changed dramatically. This seems unlikely - I measured the PZT capacitance with the BK Precision LCR meter and found it to be 2.62 nF, which is in excellent agreement with the numbers from elogs 3640 and 4354 - but this makes me wonder how the old setup ever worked. If the PZT capacitance were indeed that value, then for the Pomona box design in elog 4354, and assuming the PM at ~216kHz which was the old modulation frequency was ~30rad/V as suggested by the data in this elog, we would have had a modulation depth of 0.75 if the Function Generator were set to output a Signal at 2Vpp (2Vpp * 0.5 * 0.05 * 30rad/V = 1.5rad pp)! Am I missing something here?

Instead of using an attenuator, we could instead change the capacitor in the pomona box from 47pF mica to 5pF mica to realize a modulation depth of ~0.2 at the new modulation frequency of 231.25 kHz. In any case, as elog 4354 suggests, the phase introduced by this high-pass filter is non-zero at the modulation frequency, so we may also want to install an all-pass filter which will allow us to control the demodulation phase. This should be easy enough to implement with an Op27 and passive components we have in hand...

 

  11961   Fri Jan 29 14:43:47 2016 SteveUpdateGreen LockingInnolight laser is 10 years old
Quote:

After adjusting the alignment of the two beams onto the PD, I managed to recover a stronger beatnote of ~ -10dBm. I managed to take some measurements with the PLL locked, and will put up a more detailed post later in the evening. I turned the IMC autolocker off, turned the 11MHz Marconi output off, and closed the PSL shutter for the duration of my work, but have reverted these to their nominal state now. The are a few extra cables running from the PSL table to the area near the IOO rack where I was doing the measurements from, I've left these as is for now in case I need to take some more data later in the evening...I

Innolight 1W 1064nm, sn 1634 was purchased in 9-18-2006 at CIT. It came to the 40m around 2010

It's diodes should be replaced, based on it's age and performance.

RIN and noise eater bad. I will get a quote on this job.

The Innolight Manual frequency noise plot is the same as Lightwave' elog 11956

Attachment 1: inno1W.pdf
inno1W.pdf
  11962   Fri Jan 29 16:55:27 2016 ranaUpdateGreen LockingInnolight laser is 10 years old

I don't think there's any evidence that the noise eater is bad. That would change the behavior of the relaxation oscillation which is at 1 MHz ?

  11963   Sat Jan 30 00:12:22 2016 gautamUpdateGreen LockingInnolight laser is 10 years old

 

Quote:

I don't think there's any evidence that the noise eater is bad. That would change the behavior of the relaxation oscillation which is at 1 MHz ?

While I was investigating the AM/PM ratio of the Innolight, I found that there was a pronounced peak in the RIN at ~400kHz, which did not change despite toggling the noise eater switch on the front panel (see plot attached). The plot in the manual suggests the relaxation oscillations should be around 600kHz, but given that the laser power has dropped by a factor of ~3, I think it's reasonable that the relaxation oscillations are now at ~400kHz? 

Attachment 1: RIN_comparison.pdf
RIN_comparison.pdf
  11964   Sat Jan 30 09:56:24 2016 KojiUpdateGreen LockingInnolight laser is 10 years old

It is strange that there is no difference between with and without NE, isn't it?

  11967   Mon Feb 1 15:16:28 2016 gautamUpdateGreen LockingInnolight laser is 10 years old

The Innolight laser control unit has a 25 pin D-sub connector on the rear which is meant to serve as a diagnostics aid, and the voltages at the various pins should tell us the state of various things, like the diode power monitor, laser crystal TEC error temperature, NE status etc etc. Unfortunately, I am unable to locate a manual for this laser (online or physical copy in the filing cabinets), so the only thing I have to go on is a photocopied page that Steve had obtained sometime ago from the manual for the 2W NPRO. According to that, Pin 1 is "Diode laser 1, power monitor, 1V/W". The voltage I measured (with one of the 25 pin breakout boards and a DMM) is 1.038V. I didn't see any fast fluctuations in this value either. It may be that the coefficient indicating "normal" state of operation is different for the 1W model than the 2W model, but this measurement suggests the condition of the diode is alright after all?

I also measured the voltage at Pin 12, which is described in the manual as "Noise Eater, monitor". This value was fluctuating between ~20mV and ~40mV. Toggling the NE switch on the front of the control unit between ON and OFF did not change this behaviour. The one page of the manual that we have, however, doesnt provide any illumination on how we are supposed to interpret the voltage measured at this pin...

  11968   Mon Feb 1 15:43:18 2016 KojiUpdateGreen LockingInnolight laser is 10 years old

This is the same one as what you got from Steve. But you can find full pages.

https://wiki-40m.ligo.caltech.edu/PSL/NPRO

  11987   Fri Feb 12 11:10:49 2016 SteveUpdateGreen LockingInnolight laser is 10 years old

It shipped out for repair evaluation.

Arrived to Hayward,CA   2016Feb16

 

Attachment 1: inno1W.jpg
inno1W.jpg
  4705   Thu May 12 22:54:20 2011 ranaUpdateDAQInput Beam Naming change (no more IP)

 We decided to rename the Input Beam channels (while keeping temporary backwards compatible aliases) as:

C1:ASC-IB_POS_X, C1:ASC-IB_POS_Y, C1:ASC-IB_ANG_SUM, etc.

  2584   Tue Feb 9 17:51:48 2010 JenneSummaryIOOInput Mode Matching Telescope design is complete

The upgrade's input mode matching telescope design is complete.  A summary document is on the MMT wiki page, as are the final distances between the optics in the chain from the mode cleaner to the ITMs.  Unless we all failed kindergarden and can't use rulers, we should be able to get very good mode matching overlap.  We seem to be able (in Matlab simulation land) to achieve better than 99.9% overlap even if we are wrong on the optics' placement by ~5mm.  Everything is checked in to the svn, and is ready for output mode matching when we get there.

  7012   Mon Jul 23 20:19:01 2012 LizUpdateComputer Scripts / ProgramsInput Needed (From everyone!)

The summary pages are now online (Daily Summary), and will eventually be found on the 40m Wiki page under "LOGS-Daily Summary". (Currently, the linked website is the former summary page site)

Currently, all of the IFO and Acoustic channels have placeholders (they are not showing the real data yet) and the Weather channels are not working, although the Weather Station in the interferometer room is working (I am looking into this - any theories as to why this is would be appreciated!!).

I am looking for advice on what else to include in these pages.  It would be fantastic if everyone could take a moment to look over what I have so far (especially the completed page from July 23, 2012) and give me their opinions on:

 

1.  What else you would like to see included

2.  Any specific applications to your area of work that I have overlooked

3.  What the most helpful parts of the pages are

4.  Any ways that I could make the existing pages more helpful

5.  Any other questions, comments, clarifications or suggestions

Finally, are the hourly subplots actually helpful?  It seems to me like they would be superfluous if the whole page were updating every 1-2 hours (as it theoretically eventually will).  These subplots can be seen on the July 24, 2012 page.

My email address is endavison@umail.ucsb.edu.

 Thank you!  

 

  11896   Tue Dec 22 16:23:33 2015 gautamUpdateIOOInput alignment to PMC tweaked

When I came in this afternoon, I saw that the PZT voltage to the PMC had railed. Following the usual procedure of turning the servo gain to zero and adjusting the DC offset, I got the PMC to relock, but the PMCR level was high and the alignment looked poor on the control room monitor. So I tweaked the input alignment on the PSL till I felt it was more reasonable. The view on the control room monitor now looks more like the usual state, and the "REFL (V)" field on the PMC MEDM screen now reads 0.02-0.03 which is the range I remember it being in nominally. 

  7699   Tue Nov 13 00:15:08 2012 JenneUpdateIOOInput and Output PZTs turned on

I turned on the power supplies for the output PZTs and pitch and yaw for PZT2.  This is back to the condition that we had during atmosphere alignment, so after Ayaka has finished tweaking the MC, we can have a look at alignment of the interferometer.

  8262   Fri Mar 8 20:51:00 2013 ManasaUpdateAlignmentInput beam drift - investigation **IMPORTANT**

Checking the drift in input pointing (TT2 is the main suspect)

I have centered IPPOS and the 2/3 part of IPANG that comes out of vacuum to the QPDs to see the drift in input pointing over the weekend or atleast overnight.

If anybody would be working with the IFO alignment over the weekend, do so only after recording the drift in IPANG and IPPOS or if you will be working later tonight, center them ion the QPDs before leaving.

  8267   Mon Mar 11 12:29:25 2013 ManasaUpdateAlignmentInput beam drift - weekend trend

I centered ipang and ippos on the QPDs (using only the steering mirrors) and wanted to see the drift over the weekend.

Observation
1. IPANG has drifted (QPD sum changed from -6 to -2.5); but it is still on the QPD.
2. IPPOS does not show any drift.
3. In the plot: The jump in IPANG on the left occured when I centered the beam to the QPD and that on the right is from the 4.7 earthquake and its aftershocks this morning.

Follow-up questions
1. Do we need to worry about this drift?
2. Which of the two TTs is resposible for the drift?
3. Do the TTs tend to drift in the same direction everytime?

P.S. The TTs were not touched to center on IPANG and IPPOS. The last time they were touched was nearly 6 hours before the centering. So the question of any immediate hysteresis is ruled out.

IPANG_drift.png

  8252   Thu Mar 7 18:12:03 2013 yutaUpdateAlignmentInput beam drift ~ 0.1 mrad/h in pitch

[Jenne, Manasa, Yuta]

We temporarily centered the beam on IPANG to see input pointing drift. From eyeball, drift was ~ 0.1 mrad/h in pitch.

What we did:

  1. Aligned TT1/TT2 and aligned input pointing to Yarm.

  2. Tweaked TT2 in pitch to center the beam on the first steering mirror of IPANG path. We still saw Yarm flash in higher order modes at this point. Before tweaking, the beam was hitting at the top edge.

  3. Centered the beam on IPANG QPD.

  4. Moved IPPOS first steering mirror because IPPOS beam was not on the mirror (off in yaw, on mirror edge). Also, IPPOS beam was coming out clipped in yaw.

  5. Centered the beam on IPPOS QPD. We put lens in the path to focus the beam on the QPD.

  6. Left input pointing untouched for 4 hours.

  7. Restored TT2 again. We tried to align Y arm with IPANG available, but it was not possible without touching TRY path and AS was also clipped.

Result:
  Below is the trend of IPANG sum, X, and Y. IPANG Y (IBQPD_Y) drifted by ~0.8 counts in 4 hours. IPANG is not calibrated yet, but Jenne used her eyeball to measure beam position shift on IPANG steering mirror. It shifted by ~2 mm. This means, input pointing drifts ~0.1 mrad/h in pitch.
IPangulardrift.png

Discussion:
  Compared with yaw, pitch drift is quite large considering beam size at ETMY(~5 mm). We can monitor input pointing drift in weekends get longer trend.

Note:
  - IPANG and IPPOS are both changed from the state before pumping.

  13827   Wed May 9 17:30:04 2018 gautamUpdateGeneralInput beam misaligned

There is no beam going into the IFO at the moment. There was definitely a spot on the AS camera after I restored the suspensions yesterday, as you can see from the ASDC level in Attachment #1. But at around 2pm Pacific yesterday, the ASDC level has gone to 0. I suspect the TTs. There is no beam on the REFL camera either when PRM is aligned, and PRM's DC alignment is good as judged by Oplev.

Normally, I am able to recover the beam by scanning the TTs around with some low frequency sine waves, but not today. We don't have any readback (Oplev/OSEM) of the TT alignment, and the DC bias values havent jumped abnormally around the time this happened, judging by the OUT16 monitor points (see Attachment #2). The IMC was also locked at the time when this abrupt drop in ASDC level happened. Unfortunately, we don't have a camera on the Faraday so I don't know where the misalignment is happening, but the beam is certainly not making it to the BS. All the SOS optics (e.g. BS, ITMX and ITMY) are well aligned as judged by Oplev.

Being debugged now...

Attachment 1: InputBeamGone.png
InputBeamGone.png
Attachment 2: TTpointing.png
TTpointing.png
  13828   Wed May 9 19:51:07 2018 gautamUpdateGeneralInput beam misaligned

As suspected - the problem was with the TTs. I tested the TT signal chain by driving a low frequency sine wave using AWG and looking at the signal on an o'scope. But I saw nothing, neither at the AI board monitor point, nor at the actual coil driver mon point. I decided to look at the IOP testpoints for the DAC channels, to see if the signals were going through okay on the digital side. But the IOP channels were flatlined, as viewed on dataviewer (see Attachment #1). This despite the fact that the DAC output monitor screen in the model itself was showing some sensible numbers, again see Attachment #1.

Looking at the CDS overview screen, there were no red flags. But there was a red indicator sneakily hidden away in the IOP model's CDS status screen, the "DAC" field in the state word is red. As Attachment #2 shows, a change in the state word is correlated with the time ASDC went to 0.

Note that there are also no errors on the c1lsc frontend itself, judging by dmesg. I want to do a model restart, but (i) this will likely require reboots of all vertex FEs and (ii) I want to know if any CDS experts want to sniff for clues to what's going on before a model restart wipes out some secret logfiles. I'm a little confused that the rtcds isn't throwing up any errors and causing models to crash if the values are not being written to the registers of the DAC. It may also be that the DAC card itself is dead sad. To re-iterate, all the EPICS readbacks were suggesting that I am injecting a signal right up to the IOP.

Quoting from the runtime diagnostics note:

NOTE: As V2.7, if this error is detected, the IOP will output zero values to all DAC modules, as a protective measure. Only method to clear this is to restart the IOP and all applications on that computer
Attachment 1: DACweirdness.png
DACweirdness.png
Attachment 2: DACerror.png
DACerror.png
  7210   Thu Aug 16 20:18:39 2012 YaakovUpdateSTACISInput for feedforward/feedback in the STACIS

Below is the bottom view of the geophone preamplifier and controller for the STACIS. It slides into the upper part of the STACIS, under the blue platform. The geophone signal goes in the bottom left, gets amplified, filtered, and otherwise pampered, and goes out from the bottom right. From there it goes on to the high voltage amplifier, and finally to the PZT stacks. Below right is a closer view of the output port for the preamplifier, top and bottom.

SAM_0256.JPGSAM_0259.JPGSAM_0258.JPG

I suggest de-soldering and bending up the pins that carry the geophone signal (so the signals don't go directly to the high voltage amplifier), and adding the circuit below between the preamp and amplifier. The preamp connector is still attached to the high voltage amplifier connector in this setup, only the geophone signal pins are disconnected.

x-chip.png

More on the circuit and its placement:

The first op-amp is a summing junction, and the second is just a unity gain inverter so that signal doesn't go into the high voltage amplifier inverted. I tested this with the breadboard, and it seems to work fine (amplitudes of two signals add, no obvious distortion). The switches allow you to choose local feedback, external feedforward, or both.

The geo input will be wires from the preamp (soldered to where the pins used to go), and the external input will be BNC cables, with the source probably a DAC. The output will go to the bent up pins that used to be connected to the preamp (they go into the high voltage amplifier). This circuit can sit outside of the STACIS- there is a place to feed wires in and out right near where the preamplifier sits. For power, it can use the STACIS preamp supply, which is +/- 15V. The resistors I used in the breadboard test were 10 kOhm, and the op-amp I used was LT1012 (whose noise should be less than either input, see eLog 7190).

This is visually represented below, with the preamp pin diagram corresponding to the soldering points with the preamp upside down (top right picture):

SAM_0266.JPGSAM_0265.JPG

 

  5438   Fri Sep 16 17:16:15 2011 JenneUpdateSUSInput matrix diagonalization: Fail!

[Jenne, Anamaria]

I put the new matricies in from the free swinging test for the: ITMX, ITMY, ETMX, ETMY, PRM, BS

Some of the optics damped okay, but ETMX and BS were not good at all.  ETMX was ringing up when I turned on the damping.  BS wasn't, but when I gave it a kick, it wouldn't damp.  No good.

I tried ITMY, and it was totally fine, with nice damping Qs of ~5.  So, I don't know what's going on. 

Anamaria is trying a new 4x4 matrix-inverter, so we can look at the inversion of just the face osems.  We'll see how it goes. 

Since things were crappy, I did a BURT restore, so things are as they were earlier this morning.

  4653   Fri May 6 15:42:55 2011 valeraMetaphysicsIOOInput mode cleaner length and 11 MHz modulation frequency

 After Kiwamu set the REFL11 phases in the PRMI configuration (maximized PRM->REFL11I reesponse) I tried to measure the MC length and the 11 MHz frequency missmatch by modulating the 11 MHz frequency and measuring the PM to AM conversion after the MC using the REFL11Q signal. The modulation appears in the REFL11Q with a good snr but the amplitude does not seem to go through a clear minimum as the 11 MHz goes through the MC resonance.

We could not relock the PRMI during the day so I resorted to a weaker method - measuring the amplitude of the 11 MHz sideband in the MC reflection (RF PD mon output on the demod board) with a RF spectrum analyzer. The minimum frequency on the IFR is 11.065650 MHz while the nominal setting was 11.065000 MHz. The sensitivity of this method is about 50 Hz.

  15567   Thu Sep 10 15:43:22 2020 JonUpdateBHDInput noise spectra for A+ BHD modeling

As promised some time ago, I've obtained input noise spectra from the sites calibrated to physical units. They are located in a new subdirectory of the BHD repo: A+/input_noises. I've heavily annotated the notebook that generates them (input_noises.ipynb) with aLOG references, to make it transparent what filters, calibrations, etc. were applied and when the data were taken. Each noise term is stored as a separate HDF5 file, which are all tracked via git LFS.

So far there are measurements of the following sources:

  • L1 SRCL
  • H1 SRCL
  • L1 DHARD PIT
  • L1 DSOFT PIT
  • L1 CSOFT PIT
  • L1 CHARD PIT

These can be used, for example, to make more realistic Hang's bilinear noise modeling [ELOG 15503] and Yehonathan's Monte Carlo simulations [ELOG 15539]. Let me know if there are other specific noises of interest and I will try to acquire them. It's a bit time-consuming to search out individual channel calibrations, so I will have to add them on a case-by-case basis.

  15181   Fri Jan 31 16:04:30 2020 gautamUpdateIOOInput pointing drift

One factor which hampers locking efforts is the apparent drift of the input beam into the IFO. Over timescales of ~1 hour, I have noticed that the spot on the AS camera drifts significantly (~1 spot size) in pitch. The IPPOS QPD bears out this observation, see Attachment #1. The IMC WFS control signals do not show a correlated drift, hence my claim that the TTs are to blame.

I am able to correct this misalignment by moving TT1 in pitch (see Attachment #2, which shows some signals from a ~1 hour PRMI lock, during which time the pointing drifted, and I corrected it by moving TT1 pitch). Assuming the problem is purely TT1 pitch drifting, this corresponds to 3mm / 6m ~500urad of shift in 1 hour - seems very large. The fact that the drift is only present in pitch and doesn't really show up in yaw makes me think the problem is likely mechanical (unless the voltage to the top two coils is drifting relative to the bottom, but no LR drift, which would be very coincidental). At the moment, this is just an annoyance, but it'd be good for this problem to be fixed.

I haven't yet figured out how to make ndscope export these plots to SVG preserving the dark color theme, hence the weird light axes...

Attachment 1: IPdrift.pdf
IPdrift.pdf
Attachment 2: IPdrift_PRMI.pdf
IPdrift_PRMI.pdf
  15841   Wed Feb 24 12:29:18 2021 gautamUpdateGeneralInput pointing recovered

While working at the LSC rack, I lost the input pointing into the IFO (the TT wiring situation is apparently very fragile, and this observation supports the hypothesis that the drifting TTs are symptomatic of some electronics issue). After careful beam walking, it was recovered and the dither alignment system was used to maximize TRX/TRY once again. No lasting damage done. If I can figure out what the pin-mapping is for the TT coils in vacuum, I'm inclined to try installing the two HAM-A coil drivers to control the TTs. Does anyone know where I can find said pin-out? The wiki page links seem broken and there isn't a schematic available there...

Ok it should be possible to back it out from the BOSEM pin out, and the mapping of the in-vacuum quadrupus cable, though careful accounting of mirroring will have to be done... The HAM-A coil driver actually already has a 15 pin output like the iLIGO coil drivers that are currently in use, but the pin mapping is different so we can't just replace the unit. On the bright side, this will clear up 6U of rack space in 1Y2. In fact, we can also consider hooking up the shadow sensor part of the BOSEMs if we plan to install 2 HAM-A coil drivers + 1 Dual satellite amplifier combo (I'm not sure if this number of spares is available in what we ordered from Todd).

  3114   Thu Jun 24 11:16:32 2010 Sharmila, Rana and KiwamuHowToVACInspection of the BS (sorry, no sounds)
  3093   Mon Jun 21 14:21:34 2010 Jenne, KiwamuUpdatePhotosInspection of Magnets for the TTs

Some pictures of "magnet inspection" from Picasa.

The coating of some magnets are chipped...

  3095   Mon Jun 21 20:11:21 2010 KojiUpdatePhotosInspection of Magnets for the TTs

Were these magnets chipped before the Ni plating?

RA: Yes, it looks like this is the case. We also smashed some of the magnets against a metal surface and saw that a black grime was left. We should hold the magnets with a clean teflon clamp to measure the Gauss. Then we have to wipe the magnets before installing. I share Jenne's concern about the press-fit damaging the plating and so we need to consider using using glue or the ole magnetic attachment method. We should not rely on the structural integrity of the magnets at all.

  15865   Thu Mar 4 23:57:35 2021 KojiSummaryElectronicsInspection of the new custom dsub cables

I made the inspection of the new custom DSub cables (came from Texas).

The shelled version gives us some chance to inspect/modify the internal connections. (good)
The wires are well insulated. The conductors are wrapped with the foils and then everything is in the braid tube shield. The braid is soldered on one of the connectors. (Attachment  3/4 shows the soldering of the conductor by intentionally removing one of the insulations).

It wasn't clear that if the conductors are twisted or not (probably not).

Attachment 1: 20210304235251_IMG_0527.jpg
20210304235251_IMG_0527.jpg
Attachment 2: 20210304235302_IMG_0528.jpg
20210304235302_IMG_0528.jpg
Attachment 3: 20210304235339_IMG_0529.jpg
20210304235339_IMG_0529.jpg
Attachment 4: 20210305000050_IMG_0530.jpg
20210305000050_IMG_0530.jpg
Attachment 5: 20210305000610_IMG_0531.jpg
20210305000610_IMG_0531.jpg
Attachment 6: 20210305000615_IMG_0532.jpg
20210305000615_IMG_0532.jpg
  14175   Wed Aug 22 00:22:05 2018 KojiSummaryElectronicsInspection of the possible dual backplane interfaces for Acromag DAQ

[Johannes, Koji]

We went around the LSC, PSL, IOO, and SUS racks to check how many dual backplane interfaces will be required.

Euro card modules are connected to the backplane with two DIN 41612 connectors (as you know). The backplane connectors provide DC supplies and GND connections.
In addition, they are also used for the input and output connections with the fast and slow machines.

According to the past inspection by Johannes, most of the modules just use the upper DIN41612 connector (called P1). But there are some modules exhibited the possibility of the additional use of the other connector (P2).

Tuesday afternoon Johannes and I made the list of the modules with the possible dual use. And I took a time to check the modules with DCC, Jay's schematics, and the visual inspection of the actual modules.

LSC Rack

  • Common mode servo (D040180 Rev B)
    • Schematic source D040180 Rev B D1500308
    • Assesment: Both P1 and P2 are to be connected to Acromag, but there are only a few channels on P2
    • P1: 1A-32A Digital In
    • P2: 1A-3A Analog Out (D32/33/34, SLOW MON and spare?)
            9A Digital Out for D35 (Limitter)
            10A-15A Spare
            16A Digital In (Latch Enable/Disable)
            25A, 25C  Differential Analog in (Differential offset input, indicated as "BIAS") 
  • PD Interface (D990543 Rev B)
    • Schematic source D990543 RevB
    • Assesment: No connection necessary. We don't monitor/control anything of any LSC PDs from Acromag.

PSL Rack

  • Generic DAQ Interface (D990155) - This is a DAC interface.
    • Schematic source: Jay's page D990155 Rev.B All the lines between P2 and P3 are connected.
    • Assesment: Only P2 is to be connected to Acromag.
    • P1 DAC mon -> not necessary
    • P2 A1-A16, Connected to DAC in P2-P3
  • PMC Servo
    • Schematic source: LIGO DCC D980352
    • Assesment: Only P1 (1A-9A) is to be connected to Acromag. (Just one DSub is sufficient)
    • P1 1A-9A
  • Crystal Ref (D980353)
    • Schematic source: LIGO DCC D980353
    • Assesment: Only P1 (1A-4A) is to be connected to Acromag. (Just one DSub is sufficient)
    • P1 1A-4A
  • TTFSS REV A
    • Schematic source: PNot found
    • Assesment: Probably Only P1 is sufficient. We need to analyze the board to figure out the channel assignment.

IOO Rack

  • PD Interface (D990543 Rev B)
    • Schematic source D990543 RevB
    • Assesment: Only P1 connection is sufficient.
  • Generic DAQ Interface (D990155)
    • Assesment: Remove the module. We already have the same module in PSL Rack. This is redundant.
  • Common mode servo (D040180 Rev B)
    • See above
  • Pentek Generic Input Board D020432
    • Schematic source Jay's page D020432-A
    • Assesment: No connection. There is no signal on the backplane.

SUS Rack

  • SUS Dewhitening
    • Schematic source: Jay's page D000316-A
    • Assesment: No connection.
    • We can omit Mon CHs.
    • Bypass/Inputs are already connected to the fast channels.

 