  6815   Wed Jun 13 17:39:13 2012 yutaUpdateGreen Lockingcalibrating the beatbox

[Jenne, Yuta]

We put 0 dBm sine wave to the RF input of the beatbox and linear-sweeped frequency of the sine wave from 0 to 200 MHz using network analyzer (Aligent 4395A).
(We first tried to use 11 MHz EOM marconi)

Whlile the sweep, we recorded the output of the beatbox, C1:ALS-BEATY_(FINE|COARSE)_(I|Q)_IN1_DQ. We made them DQ channels today. Also, we put gain 10 after the beatbox before ADC for temporal whitening filter using SR560s.

We fitted the signals with sine wave using least squares fit(scipy.optimize.leastsq).
Transision time of the frequency from 200 MHz to 0 Hz can be seen from the discontinuity in the time series. We can convert time to frequency using this and supposing linear sweep of the network analyzer is perfect.

Plots below are time series data of each signal(top) and expansion of the fitted region with x axis calibrated in frequency (bottom).

ALS-BEATY_COARSE_I_IN1_DQ.pngALS-BEATY_COARSE_Q_IN1_DQ.png
ALS-BEATY_FINE_I_IN1_DQ.pngALS-BEATY_FINE_Q_IN1_DQ.png


We got

C1:ALS-BEATY_COARSE_I_IN1_DQ = -1400 sin(0.048 freq + 1.17pi) - 410
C1:ALS-BEATY_COARSE_Q_IN1_DQ = 1900 sin(0.045 freq + 0.80pi) - 95

C1:ALS-BEATY_FINE_I_IN1_DQ = 1400 sin(0.89 freq + 0.74pi) + 15
C1:ALS-BEATY_FINE_Q_IN1_DQ = 1400 sin(0.89 freq + 1.24pi) - 3.4

(freq in MHz)

The delay line length calculated from this fitted value (supposing speed of signal in cable is 0.7c) is;

  D_coarse = 0.7c * 0.048/(2*pi*1MHz) =  1.6 m
  D_fine = 0.7c * 0.89/(2*pi*1MHz) = 30 m

So, the measurement look quite reasonable.

FINE signals looks nice because we have similar response with 0.5pi phase difference.
For COARSE, maybe we need to do the measurement again because the frequency discontinuity may affected the shape of the signal.

  6814   Wed Jun 13 11:19:05 2012 steveUpdateSUSIDC receptacles clamped

The MC_ IDC 64 pin cables from sat. amplifiers to junction-interface-board towards  whitening - dewhitening at the back of rack 1 X 5 are finally  clamped with

All other sus cables of the same kind have the correct short latch arm to lock them in for reliable contact.

Attachment 1: IMG_1341.JPG
IMG_1341.JPG
  6813   Wed Jun 13 11:10:56 2012 ranaUpdateGreen Lockingmy first modescan (sort of)

You can easily calculate whether or not the coarse readout will work by thinking about the scan resolution you need given the ADC dynamic range and the whitening filter that you use.

  6812   Wed Jun 13 03:03:38 2012 yutaUpdateGreen Lockingmy first modescan (sort of)

Linear range df of the delay line technique is about df ~ c/(2D). So, the linear range for the fine signal(delay line length D=30m) is about 5 MHz.
Arm cavity FSR = c/(2L) = 3.7 MHz.
So, I think we need phase shifting to do mode scan for more than 2 FSRs by holding the arm length finely with fine servo.
For the coarse (D=1.5m), the linear range is about 100 MHz, so if we can do mode scan using coarse servo, it is OK.

In any case, I think it is nice to have linear signal with fixed slope even if we don't adjust the phase every time.

Quote:

 That sounds goofy.

With the delay line technique, you can get a linear signal over 50 MHz with no phase shifting. What is with all this I/Q stuff?

 

  6811   Wed Jun 13 02:24:02 2012 ranaUpdateGreen Lockingmy first modescan (sort of)

 That sounds goofy.

With the delay line technique, you can get a linear signal over 50 MHz with no phase shifting. What is with all this I/Q stuff?

  6810   Wed Jun 13 02:11:59 2012 yutaUpdateGreen Lockingmy first modescan (sort of)

I stabilized Y arm length by using only I phase coarse signal from the beat(C1:ALS-BEATY_COARSE_I_ERR).
I sweeped the arm length by injecting 0.05Hz sine wave from C1:ALS_OFFSETTER2_EXC.
Below is the plot of TRY and the error signal(ideally, Y arm length) while the sweep.

modescan20120612_1.png

I couldn't hold the arm length tight, so you can see multiple peaks close to each other.
We need to
  - adjust offsets
  - adjust rotation phase of I-Q mixing
  - adjust servo filters

to hold the length tighter.

Also, I couldn't sweep the Y arm length very much. I need to calibrate, but to do the modescan for many FSRs, we need to
  - introduce automatic phase optimizing system
There were sin/cos function in the CDS_PARTS, so I think we can feedback I_ERR to control rotation phase of I-Q mixing.

  6809   Tue Jun 12 23:18:18 2012 yutaUpdateGreen LockingI-Q signals for the beat

[Mengyao, Yuta]

Yes!! We have I-Q signals for the beat!!

What we did:
  1. Aligned Y arm to the Y end green incident beam. The transmission to the PSL was about 195 uW.

  2. Aligned IR beam to the Y arm by adjusting PZTs and got the transmission, C1:LSC-TRY_OUT ~ 0.86.

  3. Aligned green optics on the PSL table to get the beat signal. The beat was found when;

  PSL laser temperature on display: 31.41 deg C
  C1:PSL-FSS_SLOWDC = 1.43
  Y end laser "T+": 34.049 deg C
  Y end laser "ADJ": 0
  Y end laser measured temperature: 34.14 deg C
  C1:GCY-SLOW_SERVO2_OFFSET = 29950
  Y end slow servo: off (was on)

  4. Connected the beat PD output to the beatbox.

  5. Kicked ETMY position to change the cavity length and while the ringdown, we run pynds to get data. We plotted C1:ALS-BEATY_FINE_I_ERR vs C1:ALS-BEATY_FINE_Q_ERR, and C1:ALS-BEATY_COARSE_I_ERR vs C1:ALS-BEATY_COARSE_Q_ERR (below). We got nice circle as expected.

FINEIQplot20120612.pngCOARSEIQplot20120612.png

Current setup:
  Only AA filers are put between the output of the beatbox and the ADC.

beatysetup20120612.png

  6808   Tue Jun 12 20:35:46 2012 yutaUpdateGreen Lockingc1gcv recompiled

[Jamie, Yuta]

We recompiled c1gcv because the order of the channels were confusing. We found some change in the phase rotation module when we did this.

I did some cabling and checked each signals are actually going to the right channel. I labeled all the cables I know, which go into the AA chasis for ADC1 of c1ioo machine.

Below is the list of the channels. If you know anything about "unknown" channels, please let me know.

Current channel assignments for ADC1 of c1ioo machine:
  Red ones were added today. Green ones existed in the past, but channel assignment were changed.

cable

# on AA chassis name in Simulink channel name

connected
but unknown

J1A    
   
not connected J1B    
   
not connected J2 adc_1_2 C1:ALS-XARM_BEAT_DC
not connected adc_1_3 C1:ALS-YARM_BEAT_DC
connected
but unknown
J3    
   
connected
but unknown
J4    
   
connected
but unknown
J5    
   
connected
but unknown
J6    
   
connected
but unknown
J7    
   
beat Y arm fine I J8A adc_1_14 C1:ALS-BEATY_FINE_I
beat Y arm fine Q adc_1_15 C1:ALS-BEATY_FINE_Q
not connected J8B    
   
connected
but unknown
J9A    
   
not connected J9B    
   
connected
but unknown
J10    
   
connected
but unknown
J11    
   
not connected J12 adc_1_22 C1:ALS-BEATX_COARSE_I
not connected adc_1_23 C1:ALS-BEATX_COARSE_Q
not connected J13 adc_1_24 C1:ALS-BEATX_FINE_I
not connected adc_1_25 C1:ALS-BEATX_FINE_Q
beat Y arm coarse I
J14 adc_1_26 C1:ALS-BEATY_COARSE_I
beat Y arm coarse Q adc_1_27 C1:ALS-BEATY_COARSE_Q
not connected J15 adc_1_28 Broken! Don't use this!!
adc_1_29 (not broken)
not connected J16A adc_1_30 (not broken)
adc_1_31 Broken? Funny signal.
not connected J16B    
   

Memorandum for me:
  Recompiling procedure;

ssh c1ioo

rtcds make c1gcv
rtcds install c1gcv
rtcds start c1gcv

Attachment 1: c1gcv20120612-2.png
c1gcv20120612-2.png
  6807   Tue Jun 12 17:46:09 2012 JenneUpdateComputersrtcds: command found

Quote:

Quote:

We can't compile any changes to the LSC or the GCV models since Jamie's new script / program isn't found.  I don't know where it is (I can't find it either), so I can't do the compiling by hand, or point explicitly to the script.  The old way of compiling models in the wiki is obsolete, and didn't work :(

Sorry about that.  I had modified the path environment that pointed to the rtcds util.  The rtcds util is now in /opt/rtcds/caltech/c1/scripts/rtcds, which is in the path.  Starting a new shell should make it available again.

 Added TRX and TRY and POY11_I_ERR and POX11_I_ERR to the c1lsc.mdl using a new-style DAQ Channels block, recompiled, installed, started the model, all good.  Restarted the daqd on the framebuilder, and everything is green.  I can go back and get recorded data using dataviewer (for the last few minutes since I started fb), so it all looks good.

Note on the new DAQ Channels block:  Put the text block (from CDS_PARTS) at the same level as the channel you want to save, and name it exactly as it is in the model.  The code-generator will add the _DQ for you.  i.e. if you define a channel "TRY_OUT_DQ" in the lsc model, you'll end up with a channel "C1:LSC-TRY_OUT_DQ_DQ".

  6806   Tue Jun 12 17:29:28 2012 DenUpdateCDSdq channels

All PEM and IOO DQ channels disappeared. These channels were commented in C1???.ini files though I've uncommented them a few weeks ago. It happened after these models were rebuild, C1???.ini files also changed. Why?

I added the channels back. mx_stream died on c1sus after I pressed DAQ reload on medm screen. For IOO model it is even worse. After pressing DAQ Reload for C1IOO model DACQ process dies on the FB and IOO machine suspends.

I rebooted IOO, restarted models and fb. Models work now, but there might be an easier way to add channels without rebooting machines and demons.

  6805   Tue Jun 12 17:04:55 2012 steveUpdateSUSPRM oplevs servo is still bad

Quote:

Quote:

Yuta claims he fixed the PRM oplev by centering it the other day, but no one has left it on and watched it for a long while, to make sure it's okay.  We watched it now for ~2 min, and it was good, but we're leaving the oplevs off anyway for the night.  Tomorrow we should restore PRM (it's currently restored), turn on the oplevs, and let it sit to make sure it doesn't go crazy.

 

 PRM oplev servo was turned on with PITgain 0.5  and YAWgain  -0.7

Note: gain settings were PIT  1.0  and  YAW --0.5   on Jun 1, 2012 that I measured Feb 23, 2012

 It is still oscillating. Gains turned down to zero.

Attachment 1: PRMolvstillosc.png
PRMolvstillosc.png
  6804   Tue Jun 12 16:33:32 2012 steveUpdateSUSPRM oplevs servo ON for confirmation

Quote:

Yuta claims he fixed the PRM oplev by centering it the other day, but no one has left it on and watched it for a long while, to make sure it's okay.  We watched it now for ~2 min, and it was good, but we're leaving the oplevs off anyway for the night.  Tomorrow we should restore PRM (it's currently restored), turn on the oplevs, and let it sit to make sure it doesn't go crazy.

 

 PRM oplev servo was turned on with PITgain 0.5  and YAWgain  -0.7

Note: gain settings were PIT  1.0  and  YAW --0.5   on Jun 1, 2012 that I measured Feb 23, 2012

  6803   Tue Jun 12 13:49:32 2012 JamieConfigurationComputer Scripts / Programstconvert

A nicer, better maintained version of tconvert is now supplied by the lalapps package.  It's called lalapps_tconvert.  I installed lalapps on all the workstations and aliased tconvert to point to lalapps_tconvert.

  6802   Tue Jun 12 11:54:50 2012 JenneUpdateGreen Lockingc1gcv recompiled

Yuta added channels so we can get the Q phase of all the beat PDs to the c1gcv model.  I showed him how to recompile/install/start.

During the install, it couldn't find: Unable to find the following file in CDS_MEDM_PATH: LOCKIN_FILTER.adl

On all the screens (ALS and SUS), lockin parts are white.  Someone changed something, then didn't go back to fix the screens.

Otherwise, things look to be working fine.

Attachment 1: c1gcv20120612.png
c1gcv20120612.png
  6801   Tue Jun 12 11:25:05 2012 JamieUpdateComputersrtcds: command not found

Quote:

We can't compile any changes to the LSC or the GCV models since Jamie's new script / program isn't found.  I don't know where it is (I can't find it either), so I can't do the compiling by hand, or point explicitly to the script.  The old way of compiling models in the wiki is obsolete, and didn't work :(

Sorry about that.  I had modified the path environment that pointed to the rtcds util.  The rtcds util is now in /opt/rtcds/caltech/c1/scripts/rtcds, which is in the path.  Starting a new shell should make it available again.

  6800   Tue Jun 12 02:09:43 2012 JenneUpdateComputersmini boot fest

Quote:

stock-vector-wine-icon-on-computer-keyboard-original-illustration-58731781.jpg

 I'm starting to feel like a wine-o here.  Yuta wanted to glance at the PRM oplev dataviewer, and lo and behold, the fb lost connection just as he decided to do that.  We had checked the front end status screen not 1 minute beforehand, and everything was green.  Lame.

  6799   Tue Jun 12 02:07:42 2012 JenneUpdateSUSPRM oplevs left off

Yuta claims he fixed the PRM oplev by centering it the other day, but no one has left it on and watched it for a long while, to make sure it's okay.  We watched it now for ~2 min, and it was good, but we're leaving the oplevs off anyway for the night.  Tomorrow we should restore PRM (it's currently restored), turn on the oplevs, and let it sit to make sure it doesn't go crazy.

 

  6798   Tue Jun 12 01:58:33 2012 yutaUpdateGreen Lockingaligned Y arm to Y end green

[Jenne, Yuta]

We aligned Y arm to the Y end green incident beam.
We noticed two TEM00, bright and dim, so we decreased Y end laser temperature to 34.13 deg C.
It doubled the transmission of the green, and now the transmission to the PSL table is 178 uW, which is close to the maximum(197 uW) we got so far.

Current settings for Y end laser is;

  Y end laser "T+": 34.049 deg C
  Y end laser "ADJ": 0
  Y end laser measured temperature: 34.13 deg C
  C1:GCY-SLOW_SERVO2_OFFSET = 31025
  Y end slow servo: on (was off)

We aligned IR beam to the Y arm by mostly adjusting PZTs and got the transmission, C1:LSC-TRY_OUT ~ 0.9.

We tried to calculate the mode-matching ratio for IR by taking TRY data while ITMY and ETMY are swinging (without ALS), but it was difficult because we see too many higher order modes.

Tomorrow, we will (1) connect the beatbox to ADC, (2) edit c1gcv model, (3) scan the arm using I-Q signals.

  6797   Tue Jun 12 01:03:18 2012 JenneUpdateComputersmini boot fest

As usual, we noticed the frame builder wasn't connecting happily with the rest of the computers just as we were about to lock some stuff (we never notice it being bad when we're not trying to use the frame builder....)

All the big rectangles by each computer were white.  I restarted daqd, and that brought most things back.  c1lsc and c1sus needed their mx_streams restarted manually to get everything green again.

stock-vector-wine-icon-on-computer-keyboard-original-illustration-58731781.jpg

I've had about enough whine with my computers for tonight.

  6796   Tue Jun 12 00:15:51 2012 JenneUpdateComputersrtcds: command not found

We can't compile any changes to the LSC or the GCV models since Jamie's new script / program isn't found.  I don't know where it is (I can't find it either), so I can't do the compiling by hand, or point explicitly to the script.  The old way of compiling models in the wiki is obsolete, and didn't work :(

This means we can't (a) record TRY or (b) add the Q quadrature of the beat PD to the real time system tonight.

We're going to try just using Yuta's pynds script to capture data in real time, so we can keep working for tonight.

  6795   Mon Jun 11 22:51:11 2012 JenneUpdateComputerstdsavg not working

tdsavg isn't working:

controls@rossa:/opt/rtcds/caltech/c1/scripts/LSC 6$ tdsavg 10 C1:LSC-ASDC_IN1
ERROR: LDAQ - Unable to find NDS host "fb0"
ERROR: LDAQ - Unable to find NDS host "fb1"
ERROR: LDAQ - Unable to open socket to NDS.

 

When this command is executed inside a script, it doesn't return anything.  eg:

set offset = `tdsavg 10 C1:LSC_ASDC_IN1`
echo $offset

returns a blank line. 

Past elog research said lots of things about test points.  I didn't suspect that, since there aren't many test points occupied (according to the CDS status screens), but I cleared the test points anyway (elog 6319).  Didn't change anything, still broken.

LSCoffsets script, and any others depending on tdsavg will not work until this is fixed.

  6794   Mon Jun 11 21:50:08 2012 yutaUpdateGreen Lockingbeatbox looks OK

Summary:
  We need I-Q frequency deiscriminator to control the arm length fine and continuously.
  I checked the beatbox (LIGO-D1102241-v4; see elog #6302) and it was working.

What I did:
  1. Measured some transferfunctions with a network analyzer (Aligent 4395A) and checked the cabling is correct.

  2. Put 30 m/1.5 m delay line and checked I-Q outputs are actually orthogonal. I did this by sweeping the frequency of RF input to the beatbox. See attached picture. You can see nice circle on the oscilloscope.

Some measurement results:

  - Gains of the transferfunctions(@ 10-100MHz) between;

   RF in -> RF mon: -25 to -20 dB
   RF in -> fine delay out: -50 to -40 dB
   RF in -> coarse delay out: -50 to -40 dB
   RF in -> LO of mixer RMS-1: ~ +4 dB  (RMS-1 needs +7 dB LO)
 
  - 30m delay line(RG-142B/U) had -2 dB loss.

Note:
  - RF input must be larger than about -3 dBm to get enough LO to the mixer. Otherwise, you won't get I-Q outputs.
  - The comparator, whitening filter and differential DAQ outputs are not installed in the current beatbox.
  - Current beatbox only has electronics for the one arm.
  - The print on the board D1102241 says +15V and -15V, but they are actually opposite. Cabling is swapped in order to supply correct power to the ICs.

Attachment 1: CIMG1522.JPG
CIMG1522.JPG
  6793   Mon Jun 11 21:35:55 2012 JenneUpdateEnvironmentRattling in the HEPA

Quote:

There is an intermittent rattling sound coming from the HEPA in the NE corner of the PSL table (right above the PMC, all of our input optics).

Steve says it might be a bad bearing, but he'll check it out in the morning and get it fixed.

 MC was having a hard time staying locked, with no discernable reason from the control room (i.e. no big seismic, no PMC PZT railing).    The HEPA was on 100%, so I turned it down to 50% to hopefully reduce the rattling, if that was what was wrong. 

  6792   Mon Jun 11 16:08:58 2012 JenneUpdateEnvironmentRattling in the HEPA

There is an intermittent rattling sound coming from the HEPA in the NE corner of the PSL table (right above the PMC, all of our input optics).

Steve says it might be a bad bearing, but he'll check it out in the morning and get it fixed.

  6791   Mon Jun 11 09:37:16 2012 steveUpdateIOOPMC locked

Quote:

Quote:

IOO Angle & IOO Position QPDs centered.

 PMC trend of 400 and 1200 days

The Innolight 2W based PSL- IOO was implemented in the ~ summer of 2010

 The PMC was locked and the MC followed intantly

Attachment 1: PMClocked.png
PMClocked.png
  6790   Fri Jun 8 15:24:08 2012 steveUpdateVACRGA scan at day 276 and pressure plots

Quote:

We tried not to open chambers above 10,000 particles of 0.5 micron cf/min

New items going in:                       2 rasor beam traps, 5 badly oxidized old silver plated  setscrew with spring loaded tips......to be replaced in the future, viton tips for eq screws....some are lose, gold plated allen wrench installed at ITMX bottom, reglued magnet on ITMX

Bad hardware things found:          nylon ball "locking elements" on OSEM locking set screws with screwdriver  slot, lose  1064 nm filter on OSEM pd

 40m - pumpdown 71- maglev pumping speed - at day 276........all normal

Attachment 1: 40m71pdM276d.png
40m71pdM276d.png
Attachment 2: pd71m276d.png
pd71m276d.png
Attachment 3: pd71d276.png
pd71d276.png
  6789   Fri Jun 8 15:08:27 2012 yutaUpdateGreen Lockingaligned/mode-matched Y green beat setup

Laser temperature settings for Y arm green work today are;

  PSL laser temperature on display: 31.38 deg C (PSL HEPA 100%)
  C1:PSL-FSS_SLOWDC = 1.68
  Y end laser "T+": 34.049 deg C
  Y end laser "ADJ": 0
  Y end laser measured temperature: 34.13 deg C (*)
  C1:GCY-SLOW_SERVO2_OFFSET = 29845

Green transmission from Y end and PSL green power on the beat PD are;

  P_Y = 28 uW
  P_PSL = 96 uW

P_Y decrease from its maximum we got (75 uW, see elog #6777) is because the alignment for Y arm green is decreased. I can see the decrease from the green reflection on ETMT camera, but I will leave it because we already have enough beat.

I aligned PSL optics, including the mode-matching lens to maximize the beat note. The beat note I got is about 26dBm.
The calculated value is -14 dBm, so we have about 75 % loss.
I measured the reflection from the PD window and its reflectivity was about 30%. We still have unknown 45% loss.

  6788   Thu Jun 7 18:46:13 2012 yutaUpdateSUSPRM oplev centered

PRM oplev beam was not hitting on the QPD since Jun 1, so I centered it. I reverted the oplev servo gains and now oplev servo looks fine.

C1:SUS-PRM_OLPIT_GAIN = 1.0
C1:SUS-PRM_OLYAW_GAIN = -0.7

There's SIDE to UL/UR/LL/LR coil element in PRM TO_COIL matrix. They were 0 until Mar 31, 2012, but someone changed them to -0.160. I couldn't find elog about it.
Same thing happened to BS on Mar 13, 2012 (see elog #6409), so I think somebody did the same thing to PRM.

  6787   Thu Jun 7 17:49:09 2012 JamieUpdateCDSc1sus in weird state, running models but unresponsive otherwise

Somehow c1sus was in a very strange state.  It was running models, but EPICS was slow to respond.  We could not log into it via ssh, and we could not bring up test points.  Since we didn't know what else to do we just gave it a hard reset.

Once it came it, none of the models were running.  I think this is a separate problem with the model startup scripts that I need to debug.  I logged on to c1sus and ran:

rtcds restart all

(which handles proper order of restarts) and everything came up fine.

Have no idea what happened there to make c1sus freeze like that.  Will keep an eye out.

  6786   Thu Jun 7 13:25:37 2012 KojiUpdateSUSPRM is still oscillating

Done.

C1:SUS-PRM_OLPIT_GAIN 1.0 -> 0
C1:SUS-PRM_OLYAW_GAIN -0.7 -> 0
 

Quote:

  Set the PRM OL servo gains to zero until someone can take care of this. Turning off the buttons doesn't help anything if people run the alignment scripts.

 

  6785   Thu Jun 7 13:18:42 2012 ranaUpdateSUSPRM is still oscillating

  Set the PRM OL servo gains to zero until someone can take care of this. Turning off the buttons doesn't help anything if people run the alignment scripts.

  6784   Thu Jun 7 12:49:50 2012 JamieUpdateComputerszita network configured

Added zita to the martian host table, and configured his network accordingly:

zita            A       192.168.113.217

https://wiki-40m.ligo.caltech.edu/Martian_Host_Table

  6783   Thu Jun 7 10:39:21 2012 JamieUpdateComputerspianosa dual monitor working

I decided to remove what I thought was the problematic extra nVidia video card, since there are already two DVI outputs build-in.  The card turned out to not even be nVidia, so I don't know what was going on there.

I futzed with the BIOS to configure the primary video card, which is some new Intel card.  The lucid (10.04) support for it is lacking, but it was easy enough to pull in new drivers from the appropriate Ubuntu PPA repository:

controls@pianosa:~ 0$ sudo apt-add-repository ppa:f-hackenberger/x220-intel-mesa
controls@pianosa:~ 0$ sudo apt-add-repository ppa:glasen/intel-driver
controls@pianosa:~ 0$ sudo apt-get update
...
controls@pianosa:~ 0$ sudo apt-get upgrade
Reading package lists... Done
Building dependency tree       
Reading state information... Done
The following packages will be upgraded:
  libdrm-intel1 libdrm-nouveau1 libdrm-radeon1 libdrm2 libgl1-mesa-dri libgl1-mesa-glx libglu1-mesa xserver-xorg-video-intel
8 upgraded, 0 newly installed, 0 to remove and 0 not upgraded.
Need to get 3,212kB of archives.
After this operation, 25.2MB disk space will be freed.
Do you want to continue [Y/n]?
...
controls@pianosa:~ 0$

After a reboot, both monitors came up fine.

http://www.subcritical.org/running_ubuntu_lts_on_sandy_bridge/

  6782   Thu Jun 7 09:52:05 2012 steveUpdatePEMair cond maintenance

Quote:

 

 Air conditioning maintenance is scheduled for tomorrow from 8 to 11am

 Jeff checked and  replaced filters  as needed. Job completed this morning.

Attachment 1: PEM800d.png
PEM800d.png
  6781   Thu Jun 7 08:12:20 2012 steveUpdateSUSPRM is still oscillating

Quote:

Quote:

What not to do:

The PRM oplev servo was left on and it was driving this oscillation overnight.

 Oplev servo turned off and sus damping restored. What is kicking up the PRM?

 The PRM oscillation stopped by turnig off oplev servo.

Do not turn Oplev Servo on when PRM is missaligned !

 

Attachment 1: PRMoscil.png
PRMoscil.png
  6780   Thu Jun 7 07:27:27 2012 KojiUpdateGreen Lockingimproved Y arm green alignment - even more

Yes. The end PDH servo should be checked more carefully.

The end PDH seems to have insufficient gain at around 100Hz.
The attached is the ALS noise budget calculated with the simulink model for the green locking paper.

The residual error of the end AUX laser (green) is contributing to the final ALS signal (black).
In particular, the residual AUX frequency noise, which is shown as the blue, is contributing to the noise above 100Hz.

In the model I also confirmed that the servo still has a room for improvement.

By having more suppression of the AUX frequency noise, we will be able to increase the ALS servo bandwidth
without increasing the ALS residual RMS by a servo bump. This will give us the improvement of the seismic
noise suppression at 10Hz (i.e. more suppression of red).

Attachment 1: freq_noise_AUX.pdf
freq_noise_AUX.pdf
  6779   Thu Jun 7 05:39:41 2012 yutaUpdateGreen Lockingcoarsely stabilized Y arm length with ALS

I coarsely stabilized Y arm length to off resonance point for IR using ALS.
Currently, ASL servo loop is unstable and oscillates so much that I can't hold the length to the resonance point.
We need more investigation on the servo loop before doing the mode scan.

Below is a snapshot of ALS medm screens and time series data of the error signal for ALS coarse loop (C1:ALS-BEATY_COARSE_I_ERR) and IR transmission for the Y arm (C1:LSC-TRY_OUT) when I turned the servo on.

MyFirstALS20120607.png

 

Note:
  I took off amplifiers right after the beat PD on PSL table.
  Also, I reverted the gain change Jenne made last night (elog #6750), because they no longer show overload lights.

  6778   Thu Jun 7 03:37:26 2012 yutaUpdateCDSmx_stream restarted on c1lsc, c1ioo

c1lsc and c1ioo computers had FB net statuses all red. So, I restarted mx_stream on each computer.

ssh controls@c1lsc
sudo /etc/init.d/mx_stream restart

  6777   Thu Jun 7 02:59:31 2012 yutaUpdateGreen LockingY green beat - found it!!

Summary:
  I found the big green beat note for the Y arm. The alignment of the green optics on the PSL table was crappy.

What I did:
  1. By adjusting PSL laser temperature, I found tiny beat note when

  PSL laser temperature on display: 31.35 deg C (PSL HEPA 100%)
  C1:PSL-FSS_SLOWDC = 1.75

and

  PSL laser temperature on display: 33.21 deg C (PSL HEPA 100%)
  C1:PSL-FSS_SLOWDC = -6.82

Y end laser temperature settings are fixed as follows during the measurement.

  Y end laser "T+": 34.049 deg C
  Y end laser "ADJ": 0
  Y end laser measured temperature: 34.13 deg C (*)
  C1:GCY-SLOW_SERVO2_OFFSET = 29845

Bryan's formula (swapped one; see elog #6746),  suggests the paring

  (Yend laser temp, PSL laser temp) = (34.13 deg C, 31.09 deg C).

  2. Checked that beat PD is working by swapping the beat PDs for Y arm and X arm.

  3. Checked that the mode-matching of the two beams, one from Y arm and the other from PSL, is OK by moving mode-matching lens and measuring the beam spot size at near/far field are the same.

  4. When checking the beam spot size at far field(~ 1 m from the BS), I noticed the relative beam tilt by ~ 1 mrad. We aligned them few days ago, but I think the green beam from the Y arm has shifted. Of course we align IR to the Y arm first, but we difinitely need dither servo or A2L for the arm, too.

  5. As soon as aligning the PSL green optics near the BS, I found a large beat note. The measured amplitude was ~ -26 dBm, without any amplifiers after the PD.

  Currently the measured green beam power onto the beat PD from Y end is 75 uW and from PSL is 92 uW. So the calculated beat amplitude will be ~ -10 dBm (see calculation in elog #6746). So there is about 84% loss. Anyway, I will go on to the mode scan.

  6776   Thu Jun 7 02:25:27 2012 yutaUpdateGreen Lockingimproved Y arm green alignment - even more

[Koji, Yuta]

Summary:
  We improved the Y arm green transmission to the PSL table. It is now 197 uW.
  The improvement was done mainly by adjusting the Y arm green servo gain.

What we did:
  1. Fine-adjusted steering mirrors after the faraday on Y end table by monitoring Y arm green transmission (used Thorlabs PDA36A as a PD, C1:GCV-GREEN_TRY as a channel). We decided which way to adjust the mirrors by just pushing/pulling its mount.

  2. The output of the reflection PD on the oscilloscope seemed like the Y end frequency servo was oscillating. So, we reduced the amplitude of the frequency modulation from 2.83 V to 0.13 V.

  3. We noticed there were two TEM00, one is brighter and the other is dim. We thought this came from a mode-hopping or something. So, we changed the Y end laser temperature from 34.68 deg C to 34.13 deg C (measured). This reduced dim TEM00 and the main one got brighter. C1:GCY-SLOW_SERVO2_OFFSET was changed from 29425 to 29845.

  4. Fine-adjusted the position of the mode-matching lens by reduing LG modes.

Current green power:
  Current measured green power values are as follows.

beatygreenpower20120607.png

  Calculated value for the Y arm green transmission is ~ 600 uW, but we think we are almost at the maximum we can get. So, we have about 70% loss from the Y end table to the PSL table. There may be large loss in windows. The beam shape of the transmitted beam seems OK, but there may be some clipping.

To do:
  - Fine tune the Y end frequency servo loop. Reducing the amplitude of the frequency modulation for reducing the gain is not a very good idea.

  6775   Thu Jun 7 01:46:05 2012 yutaSummaryGreen LockingY green beat - found it!!

I found the big big Y green beat. Details will be posted later.

CIMG1504.JPG

  6774   Wed Jun 6 22:25:05 2012 JamieUpdateComputerszita now running ubuntu 10.04

I'll finish the rest of the config to turn it into a full-fledged workstation tomorrow.

  6773   Wed Jun 6 22:23:53 2012 JamieUpdateComputersFailed attempt to get pianosa to support dual monitors

I managed to get pianosa working again with just a single monitor.  I'm done trying to configure it for dual, though.

  6772   Wed Jun 6 22:04:19 2012 JamieUpdateComputersFailed attempt to get pianosa to support dual monitors

I've spent an inordinate amount of time trying to figure out how to get pianosa to support dual monitors.  It apparently has some special nvidia graphics that are not well supported in Ubuntu (10.04 at least).  I've tried installing a newer kernel (3.0.0) and installing the special nvidia compile-from-source Ubuntu packages, but nothing is working.  And now unfortunately he's not giving me any video at all.  I'm sick of this today, so I'll try again tomorrow.

In the future we should avoid these bullshit nvidia cards like the plague.

  6771   Wed Jun 6 21:11:45 2012 yutaUpdateGreen Lockingimproved Y arm green alignment

Summary:
  Y arm green transmission to the PSL table improved from ~ 20 uW to 61 uW. Improvement was done by adjusting steering mirrors before and after the faraday on the Y end table.
  But 61 uW is not enough!

What I did:
  1. The incident power to the faraday for the green beam on the Y end table was 1.4 mW, but the transmission was 1.2 mW. So, I adjusted the steering mirrors and the transmission increased to 1.4 mW.

  2. I found that adjusting the steering mirrors to the faraday also increased alignment of the green beam to the Y arm. We always adjusted only the steering mirrors after the faraday for the alignment. I adjusted the alignment using both steering mirrors this time. Reflection of the green beam on the ETMYT camera seems more reasonable now and more frequently lock to TEM00 when closing and opening the Y end green servo loop.

  3. For the adjustment, I tried to utilize PD at the reflection port, or the transmission port. However, I couldn't do that because they fluctuates too much. I don't know why.

  4. Measured the green transmission to the PSL table, The transmitted power was ~20 uW, but after the aligning, it improved to 61 uW.

Current green power:
  I measured the green beam power at various places using Newport power meter (Model 840) with its filter on.

beatygreenpower20120606.png

  Incident green power to the Y arm is ~ 1 mW (more than 1 mW because the aparture of the power meter was smaller than the beam size) and Y arm transmission is designed to be 55%. So, if the alignment and mode matching are perfect, the transmission to the PSL table should be ~ 600 uW. The measured value 61 uW seems too small. Kiwamu says it was 140 uW when he did Y arm.

Next:
  I will find the beat note again tonight and check if the beat PD is working correctly and if the mode matching of the two beams at the PSL table is good.

  6770   Wed Jun 6 19:46:46 2012 SureshSummaryIOOTip-tilt assembly: current status and work remaining

 

Recent History

The lower blades which I had given to the Physics Workshop for making a vacuum relief hole (using a sinker-EDM process) came back about ten days ago.   Merih Eken <meken@caltech.edu>,  the supervisor at the Physics Dept workshop, handled this matter for us.  The blades were sent to a local EDM machineshop and returned in about three working days ( a weekend intervened). 

IMG_0685.JPG  IMG_0687.JPG

Bob cleaned and handed them over to me yesterday evening.  

Current status

Today I have reassembled the four tip-tilts.  I have repacked them in clean bags (double bagged) shifted them to Clean Optics Cabinet (near the ETMX chamber).  The four tip-tilts are in the bottom-most shelf in the cabinet.  There are also some tip-tilt spares in a separate envelope.

Note:  The mirror holder is now held tightly by the eddy current dampers.  This was done for safety of the wires during transportation from LHO.  The eddy current damper in the front of the mirror has to be retracted to allow the mirror holder to swing free.  It has be to about 1mm away from the suspended mirror holder

Work Remaining

1) We need to install the quadrapus cables.  The connector placement on the BOSEM side will have some issues.  It is best to loosen the BOSEM seating as well as the connector seating screws and then push the cable connector into place.  Caution:  when the connector seating screws on the BOSEM are loosened the flexible ckt could be damaged by the loose connector.

2) Insert the mirrors into the mirror holders and balance the suspension such that the mirror's hang vertical and do not have a large yaw offset.

3) Adjust the wire suspension point height so that the flags are in the center of the BOSEM aperture.  Else they will strike against the

4) We need to adjust the position of the BOSEMs such that the shadow sensor signals are at 50%.  This ensures that all the magnets hang at an appropriate distance from their respective coils.

5) To do (3) we need to set up a shadow sensor read-out set-up for one tip-tilt (four sensors)

 

Attachment 2: IMG_0687.JPG
IMG_0687.JPG
  6769   Wed Jun 6 18:22:52 2012 JamieUpdateComputer Scripts / Programshacked ezca tools

Quote:

I didn't like this solution, so I hacked up something else.  I made a new single wrapper script to handle all of the utils.  It then executes the correct command based on the zeroth argument (see below).

I think moved all the binaries to give them .bin suffixes, and the made links to the new wrapper script.  Now everything should work as expected, with this new retry feature.

Yuta and I added a feature such that it will not retry if the environment variables EZCA_NORETRY is set, e.g.

$ EZCA_NORETRY=true ezcaread FOOBAR

  6768   Wed Jun 6 18:04:22 2012 JamieUpdateComputer Scripts / Programshacked ezca tools

Quote:

Currently, ezca tools are flakey and fails too much.
So, I hacked ezca tools just like Yoichi did in 2009 (see elog #1368).

For now,

/ligo/apps/linux-x86_64/gds-2.15.1/bin/ezcaread
/ligo/apps/linux-x86_64/gds-2.15.1/bin/ezcastep
/ligo/apps/linux-x86_64/gds-2.15.1/bin/ezcaswitch
/ligo/apps/linux-x86_64/gds-2.15.1/bin/ezcawrite

are wrapper scripts that repeats ezca stuff until it succeeds (or fails more than 5 times).

Of course, this is just a temporary solution to do tonight's work.
To stop this hack, run /users/yuta/scripts/ezhack/stophacking.cmd. To hack, run /users/yuta/scripts/ezhack/starthacking.cmd.

Original binary files are located in /ligo/apps/linux-x86_64/gds-2.15.1/bin/ezcabackup/ directory.
Wrapper scripts live in /users/yuta/scripts/ezhack directory.

I wish I could alias ezca tools to my wrapper scripts so that I don't have to touch the original files. However, alias settings doesn't work in our scripts.
Do you have any idea?

I didn't like this solution, so I hacked up something else.  I made a new single wrapper script to handle all of the utils.  It then executes the correct command based on the zeroth argument (see below).

I think moved all the binaries to give them .bin suffixes, and the made links to the new wrapper script.  Now everything should work as expected, with this new retry feature.

controls@rosalba:/ligo/apps/linux-x86_64/gds-2.15.1/bin 0$ for pgm in ezcaread ezcawrite ezcaservo ezcastep ezcaswitch; do mv $pgm{,.bin}; ln ezcawrapper $pgm; done
controls@rosalba:/ligo/apps/linux-x86_64/gds-2.15.1/bin 0$ cat ezcawrapper
#!/bin/bash

retries=5

pgm="$0"
run="${pgm}.bin"

if ! [ -e "$run" ] ; then
    cat <&2
This is the ezca wrapper script.  It should be hardlinked in place of
the ezca commands (ezcaread, ezcawrite, etc.), and executing the
original binaries (that have been moved to *.bin) with $retries
failure retries.
EOF
    exit -1
fi

if [ -z "$@" ] || [[ "$1" == '-h' ]] ; then
    "$run"
    exit
fi

for try in $(seq 1 "$retries") ; do
    if "$run" "$@"; then
	exit
    else
	echo "retrying ($try/$retries)..." >&2
    fi
done
echo "$(basename $pgm) failed after $retries retries." >&2
exit 1

  6767   Wed Jun 6 15:16:00 2012 yutaUpdateIOOMC WFS offsets adjusted

MC reflection (C1:IOO-MC_RFPD_INMON) got worse when WFS servos were on. After aligning MC optics, it will be ~0.5 but if I turned on WFS, it became ~0.8.
I measured the beam spot positions on MC optics. They seemed like the same from the measurement yesterday.

# filename      MC1pit  MC2pit  MC3pit  MC1yaw  MC2yaw  MC3yaw  (spot positions in mm)
./dataMCdecenter/MCdecenter201206052111.dat     3.234388        4.234564        2.654212        -6.656221       -0.677541       4.506170       
./dataMCdecenter/MCdecenter201206061420.dat     3.300867        4.567555        2.692971        -6.484464       -1.705443       4.423250

So, I ran /opt/rtcds/caltech/c1/scripts/MC/WFS/WFS_FilterBank_offsets to adjust the WFS offsets.

C1:IOO-MC_RFPD_INMON is now ~ 0.5 and  C1:IOO-MC_TRANS_SUM is now ~ 2.7e3 with WFS on.

  6766   Wed Jun 6 14:36:58 2012 steveUpdateSUSETMY oplev work finished

Quote:

Quote:

Quote:

 

 The typical sign of a dying gas laser is that it glows for a few minutes only. The power supplies are fine.

Two new  JDS - Uniphase 1103P lasers ( NT64-104 )  arriving on Monday, May 21

 Yesterday I swapped in new He/Ne laser with output power 3.5 mW  The return spot on qpd is large ~6mm in diameter and 20,500 counts

The spot size reduction require similar layout as ETMX oplev.

 The oplev path is relayed and the spot size on the qpd is reduced. I still have to clean up and replace "Miki Mouse" lens holder.

 Late entry for Monday morning June 4, 2012

Cables were stress relieved. Cable entry - exit ports enlarged. Air gaps were minimized.