  6604   Sat May 5 01:24:07 2012 DenUpdatePSLPMC

I was interested what whitening filter do we have between MC servo and ADC. The shape is in the figure below, SR provided 1 V white noise. Before the whitening filter MC_F is measured in Volts with SR and ADC (for ADC the shape is calculated using the whitening filter form):

whitening.jpg     mcf_v.png

I also wondered if FSS or PZT servo can add noise to the mode cleaner length signal and what is their gain. It should be big, as the laser's calibration is ~1 MHz/V => to account for seismic noise of 10^-6 m at 1 Hz, the voltage given to the laser should be ~ 1 V. And it is indeed the case. The gain is ~1000. I measured the coherence between MC_F and the laser fast input. It is 1 in the range measured (0.05 - 100 Hz). FSS and PZT do not add significant noise.

Unfortunately, after the measurement when I unplugged BNS connector from the laser, I misaligned PMC. For several hours I adjusted the mirrors but could not significantly improve transmitted signal. I'll return to this issue tomorrow.

  6603   Fri May 4 17:46:46 2012 JenneUpdateGreen LockingPSL doubling oven back on

While walking past the PSL, I noticed that the PSL's doubling oven's heater was still disabled from the power outage.  As with the ETMY heater, I hit the Enable button, and it started warming up (according to the front panel at least).

  6602   Fri May 4 17:44:42 2012 JenneUpdateGeneralOpLev 90-day trend

Quote:

After fixing the PRM tonight, all of the oplevs look okay. 

.....except ITMX, which looks like it's power dropped significantly after the CDS upgrade.  To be investigated tomorrow.

 I had a look-see at ITMX's oplev.  I can't see any clipping, so maybe the power is just low?  One thing that was funny though is the beam coming directly from the laser.  There is the main, regular beam, and then there is a thin horizontal line of red light also coming straight out of the laser.  I don't know what to do about that, except perhaps put an iris right after the HeNe, to block the horizontal part?  I'm not sure that it's doing anything bad to the optic though, since the horizontal part gets clipped by other optics before the beam enters the vacuum, so there is no real irregularity on the beam incident on the QPD.

I realigned the oplev on the QPD, using last night's ITMX alignment + whatever drift it picked up over night, so it may need re-recentering after Xarm is nicely aligned.

  6601   Thu May 3 22:37:44 2012 JenneUpdateGeneralOpLev 90-day trend

After fixing the PRM tonight, all of the oplevs look okay. 

.....except ITMX, which looks like it's power dropped significantly after the CDS upgrade.  To be investigated tomorrow.

Attachment 1: OpLevTrends_90days_Ending3May2012
  6600   Thu May 3 21:13:48 2012 KojiSummarySUSITMX/PRM/BS OPLEV aligned

[Jenne/Den/Koji]

We locked Xarm/Yarm and manually alignmed ITMX/ITMY/BS/ETMX/ETMY/PZT1/PZT2.

ITMY OPLEV was largely misaligned ==> The beam was centered on the QPD.

----

Then we aligned PRM using SB locking PRMI.

We noticed that OPLEV servo does not work. It made the PRM just noiser.

We went into the PRM table and found that the OPLEV beam was clipped in the vacuum chamber.
We tried to maximize the reflected beam from the window by touching the steering mirrors at the injection side.

Then the reflected beam was introduced to the center of the QPD.

After the alignment, the OPLEV QPD SUM increased to 4000ish from 200ish.
According to the OPLEV trend data, this is a nominal value of the QPD SUM.

Now the OPLEV servo does not go crazy.

 --

BS OPLEV beam was centered on the QPD.

  6599   Thu May 3 19:52:43 2012 JenneUpdateCDSOutput errors from dither alignment (Xarm) script

epicsThreadOnceOsd epicsMutexLock failed.
Segmentation fault
Number found where operator expected at -e line 1, near "0 0"
        (Missing operator before  0?)
syntax error at -e line 1, near "0 0"
Execution of -e aborted due to compilation errors.
Number found where operator expected at (eval 1) line 1, near "*  * 50"
        (Missing operator before  50?)
epicsThreadOnceOsd epicsMutexLock failed.
Segmentation fault
Number found where operator expected at -e line 1, near "0 0"
        (Missing operator before  0?)
syntax error at -e line 1, near "0 0"
Execution of -e aborted due to compilation errors.
syntax error at -e line 1, at EOF
Execution of -e aborted due to compilation errors.
Number found where operator expected at (eval 1) line 1, near "*  * 50"
        (Missing operator before  50?)
epicsThreadOnceOsd epicsMutexLock failed.
status : 0
I am going to execute the following commands
ezcastep -s 0.6 C1:SUS-ETMX_PIT_COMM +-0,50
ezcastep -s 0.6 C1:SUS-ITMX_PIT_COMM +,50
ezcastep -s 0.6 C1:SUS-ETMX_YAW_COMM +,50
ezcastep -s 0.6 C1:SUS-ITMX_YAW_COMM +-0,50
ezcastep -s 0.6 C1:SUS-BS_PIT_COMM +0,50
ezcastep -s 0.6 C1:SUS-BS_YAW_COMM +0,50
hit a key to execute the commands above

  6598   Thu May 3 17:15:38 2012 KojiUpdateLSCc1iscaux2 rebooted/burtrestored

[Jenne/Den/Koji]

We saw some white boxes on the LSC screens.
We found c1iscaux2 is not running.

Once the target was power-cycled, these epics channels are back.
Then c1iscaux2 were burtrestored using the snapshot at 5:07 on 4/16, a day before the power glitch.

  6597   Thu May 3 13:35:56 2012 JenneBureaucracyGeneral40m Meeting Action Items

Action Items:

IFO:

         Arm cavity sweeps, mode scan - JENNE

        Align AS OSA (others?) - JENNE       

        Investigate PRMI glitches, instability (take PRM oplev spectra locked, unlocked, to see if PRM is really moving) - KOJI, JENNE, DEN

         Connect up beatbox for Xarm use - KOJI with JENNE       

 

TT:

Prepare electronics for TTs (coil drivers) - JAMIE

In-air TT testing to confirm we can control / move TTs before we vent (starting in ~2 weeks) - SURESH

Connect TTs to digital system and controls, lay cables if needed - JAMIE with SURESH

OAF:

OAF comparison plot, both online and offline, comparing static, adaptive and static+adaptive - DEN

Static-only OAF noise budget (Adaptive noise budget as next step) - DEN

Build amplifiers for new small microphones - DEN 

SLC:

Black glass: to clean&bake - KOJI

Scattered light measurement at the end stations: design / confirmation of the mechanical parts/optics/cameras - JAN

MM:

IPPOS beam measurement - SURESH with JENNE

AS beam measurement (if beam is bright enough) - SURESH and JENNE

Mode matching calculations, sensitivity to MC waist measurement errors, PRM position - JENNE

Think up diagnostic measurement to determine mode matching to PRC while chambers are open, while we tweak MMT - JAMIE, JENNE, KOJI, SURESH

Computers:

       Upgrade Rossa, Allegra to Ubuntu, make sure Dataviewer and DTT work - JAMIE

 

  6596   Thu May 3 13:19:17 2012 JenneUpdateLockingMore success last night

I locked each arm, and the Michelson last night, no problems after I increased the Yarm gain from 0.1 to 0.2 .  I checked the green beam alignment just before going home, and both of the green beams are locking on ~03 or 04 modes, so aligning them is on the list for today.

  6595   Thu May 3 13:07:05 2012 KojiBureaucracyGeneralInterferometer check-up session

From 2PM, Jenne/Den/Koji. Finished at 10:30PM

  • General
    • More LCD screens! Wider screens!
      • Allegra - only can handle single head right now, but should prepare for the eventual upgrade of that videocard/whole computer and get 2 monitors.
      • Every workstation should have 2 24" monitors.  We currently have 2 good ones (one on Rossa, one on Pianosa), so we need 6.
    • Remove Control room shelf currently holding OSA 'scopes
    • Adjust both TV shelves so that they're at the same height, ~6" higher than the current left shelf.
      This gives space for stacking the 24" monitors for each computer.
    • Audio system for the signals!!!! Even a crappy one!
           /cvs/cds/ligo/centos5.5/cds/project/gds-2.14.2/Monitors/rockIFO
           /cvs/cds/gds/gds/Monitors/rockIFO
           (really /cvs/cds, not /opt)

  • IOO
    • PZT or Picomotor mounts for PSL/ALS beams
    • ALS on the both arm simultaneously / common/diff ALS scripts
  • SUS
    • PRM OPLEV servos are sometimes not stable. (fixed, JCD/DM/KA, 5/2/2012)
    • ETMX oplev - replace 1064nm lens with 632nm lens (KBC037, -100)
  • IFO
    • It seems that there is an occasional common-mode power transient in the arm transmissions.
      -> Track down the cause and fix.
    • Fix arm ASS even on CentOS
    • Drift of the green incident axis -> Assess the amount of the drift / replace the mount
    • Calibration of POP22 / AS110
    • PMC/IMC/ARM characterization (loss, finesse, reflectivity, etc)
  • CDS
    • Capture OSA signals in CDS (the 'scope TDS1001B has a USB port in the back for connecting to the computer)
    • Transmon (arms) for high and low power
    • POX11 whitening is not toggling the analog whitening???
    • CentOS machines cannot open simulink models properly (get "Bad Link"s everywhere, so you can't do anything useful).
    • Dataviewer and DTT can't connect to framebuilder from CentIOS machines
  • Electronics
    • Actuator noise level characterization (coil driver response in V/m & coil driver noise level V/rtHz)
    • Beat box
    • I/Q DFD
    • Improvement of POP22/110/AS110 RF circuits.
  • MEDM
    • OL alert in the watch dog screen (this is awesome!) should have small text saying "OL" (done, JCD, 5May2012)
    • Complete 40m overview screen - everything should be clickable with pseudo 3D icons
    • BETTER SUSPENSION SCREEN!!!
    • Script to generate a MEDM tree
    • Resurrect MEDM snapshots
       
    • The LSC screen has some "white" boxes -> investigate and fix them (done, KA, 5/2/2012)
    • Make the LSC control screen (compact) in the vertical arrangement (done, KA, 5/2/2012)
    • The signals of the watch dog screen should go red if any of the WDs are not enabled.
      =>Copy the logic of the signals in LSC screen.
      (done, KA, 5/2/2012)
    • Remove c1gfd from the cds status screen (done, KA, 5/2/2012)
    • Add "BURT" status indicators on the cds status screen (done, KA, 5/2/2012)
  • SCRIPT
    • Locking scripts with integrator triggers / disabling when unlocked
    • Daily diagnosis of the MC spot positions (there must be something already...)
    • Daily/occasional adjustment of the incident axis on the MC
    • Suspension "misalign/restore" script is too wild => make a new script (Done, JCD, 7May2012)
    • IFO_CONFIGURE scripts still do a burt restore, so step the optics.  Need to remove optic alignment from config .snap files, use reg restore script instead.
    • OPLEV/OSEM trending script before the IFO work for diagnosis.
    • Auto-locker for PMC/Arm/etc
  • Video
    • If each video screen has a caption, that would be great
    • GUI interface of "videoswitch" Mike!
    • Mouse Jiggler for Zita (called Swinput?)
  • Ubuntu
    • burttoday is not aliased in ubuntu. burttoday:      aliased to cd /opt/rtcds/caltech/c1/burt/autoburt/today/
    • ASS doesn't run on Ubuntu!
      ezcawrite: error while loading shared libraries: libca.so: cannot open shared object file: No such file or directory
      ezcaswitch: error while loading shared libraries: libca.so: cannot open shared object file: No such file or directory
       
  6594   Wed May 2 21:04:06 2012 DenUpdatePEMEM 172 coherence

I measured coherence between 2 EM 172 microphones in a "quiet room" with SR785

em_coh.png

High-frequency noise (>2k) is SR785 noise - I'm not using any amplifier now, the signal from microphone is sent directly to SR785 and is weak at high frequencies.

  6593   Wed May 2 19:47:20 2012 DenUpdatePEMEM 172 nonlinearities

I've checked new small EM172 microphones for nonlinearities using Koji's Mac Book speakers. EM172 + Mac nonlinearities are presented at the figure

500Hz_noise.png

The Mac's sound frequency was specified to be 500 Hz. Harmonics are seen at 1k, 1.5k, but their amplitude is ~30 times less.

  6592   Tue May 1 17:42:15 2012 Mike J.UpdateComputersSensoray

I've upgraded the Sensoray GUI so it can now switch the video channel it receives, thanks to the videoswitch script.

V4L2_Capture_Demo_r01.png

  6591   Tue May 1 08:18:50 2012 JamieUpdateCDSFrame Builder is down

Quote:

 

I tried restarting the fb in two different ways.  Neither of them re-established the connection to dtt or epics.

 Please be conscious of what components are doing what.  The problem you were experiencing was not "frame builder down".  It was "dtt not able to connect to frame builder".  Those are potentially completely different things.  If the front end status screens show that the frame builder is fine, then it's probably not the frame builder.

Also "epics" has nothing whatsoever to do with any of this.  That's a completely different set of stuff, unrelated to DTT or the frame builder.

  6590   Mon Apr 30 22:58:57 2012 JenneUpdateElectronics11MHz Marconi set to default after power outage

After a power outage, a Marconi comes back to it's defaults.  It needed to be reset to the values in elog 5530.  I'm putting a label on the Marconi so we don't have to look it up next time.

Before fixing the Marconi, POY11, AS11 and AS55 all looked like noise, no real signals, even though the arm is flashing.  Now they all look PDH-y, so things are better.

  6589   Mon Apr 30 22:41:34 2012 JenneUpdateSUSETMY oplev power supply dead

Quote:

And has the temp controller of the green been recovered too? (Push enable button)

 

 Just now, yes.

  6588   Mon Apr 30 21:24:43 2012 KojiUpdateSUSETMY oplev power supply dead

And has the temp controller of the green been recovered too? (Push enable button)

 

  6587   Mon Apr 30 21:05:32 2012 JenneUpdateSUSETMY oplev power supply dead

[Jenne, MikeJ]

The ETMY oplev quad sum was ~0, so we went to investigate.  The power supply was dead.  Keying it on and off didn't fix things, and it was certainly getting power since the front indicator light was flickering.  We replaced it with a different power supply, and the oplev is back.

Also, the Y-green AUX laser was off, presumably from the power outage a while back.  I turned it back on.

 

EDIT JD:  The first power supply we used didn't work....the oplev was on for a few minutes, then went off again.  We switched to one of the new (still JDSU) power supplies from Edmund Optics, and it's been happily working for at least an hour now.

  6586   Mon Apr 30 20:43:33 2012 SureshUpdateCDSFrame Builder is down

Quote:

Quote:

Quote:

Frame builder is down.  PRM has tripped its watch dogs.  I have reset the watch dog on PRM and turned on the OPLEV. It has damped down.  Unable to check what happened since FB is not responding.

There was an minor earthquake yesterday morning which people could feel a few blocks away.  It could have caused the the PRM to unlock.

Jamie,Rolf,  is it okay or us to restart the FB?  

 If it's down it's alway ok to restart it.  If it doesn't respond or immediately crashes again after restart then it might require some investigation, but it should always be ok to restart it.

I tried restarting the fb in two different ways.  Neither of them re-established the connection to dtt or epics.

1) I restarted the fb from the control room console with the 'shutdown' command.  No change.

2) I halted the machine with 'shutdown -h now' and restarted it with the hardware reset button on its front-panel. No change.

The console connected to the fb showed that the network file systems did not load.  Could this have resulted in failure to start several services since it could not find the files which are stored on the network file system?

The fb is otherwise healthy since I am able to ssh into it and browse the directory structure.

[Mike, Rana]

The fb is okay.  Rana found that it works on Pianosa, but not on Allegra or Rossa.  It also works on Rosalba, on which Jamie recently installed Ubuntu.

The white fields on the medm  'Status' screen for fb are an unrelated problem.

 

 

  6585   Mon Apr 30 18:46:34 2012 ranaUpdateComputersmegatron

Last week I found that megatron was still off after the power outage. Apparently, the power recovery checklist had not been followed during the recovery.

Please remember to use the checklist and post the checklist results after each power outage. Megatron is now on and functioning.

  6584   Mon Apr 30 16:56:05 2012 SureshUpdateCDSFrame Builder is down

Quote:

Quote:

Frame builder is down.  PRM has tripped its watch dogs.  I have reset the watch dog on PRM and turned on the OPLEV. It has damped down.  Unable to check what happened since FB is not responding.

There was an minor earthquake yesterday morning which people could feel a few blocks away.  It could have caused the the PRM to unlock.

Jamie,Rolf,  is it okay or us to restart the FB?  

 If it's down it's alway ok to restart it.  If it doesn't respond or immediately crashes again after restart then it might require some investigation, but it should always be ok to restart it.

I tried restarting the fb in two different ways.  Neither of them re-established the connection to dtt or epics.

1) I restarted the fb from the control room console with the 'shutdown' command.  No change.

2) I halted the machine with 'shutdown -h now' and restarted it with the hardware reset button on its front-panel. No change.

The console connected to the fb showed that the network file systems did not load.  Could this have resulted in failure to start several services since it could not find the files which are stored on the network file system?

The fb is otherwise healthy since I am able to ssh into it and browse the directory structure.

  6583   Mon Apr 30 13:58:25 2012 JamieUpdateCDSFrame Builder is down

Quote:

Frame builder is down.  PRM has tripped its watch dogs.  I have reset the watch dog on PRM and turned on the OPLEV. It has damped down.  Unable to check what happened since FB is not responding.

There was an minor earthquake yesterday morning which people could feel a few blocks away.  It could have caused the the PRM to unlock.

Jamie,Rolf,  is it okay or us to restart the FB?  

 If it's down it's alway ok to restart it.  If it doesn't respond or immediately crashes again after restart then it might require some investigation, but it should always be ok to restart it.

  6582   Mon Apr 30 13:00:50 2012 SureshUpdateCDSFrame Builder is down

Frame builder is down.  PRM has tripped its watch dogs.  I have reset the watch dog on PRM and turned on the OPLEV. It has damped down.  Unable to check what happened since FB is not responding.

There was an minor earthquake yesterday morning which people could feel a few blocks away.  It could have caused the the PRM to unlock.

Jamie,Rolf,  is it okay or us to restart the FB?  

  6581   Fri Apr 27 13:32:06 2012 DenUpdatePEMseism channels

A few weeks ago I found that GUR2_X signal is biased from 0 to 800 counts in average. I decided that the corresponding channel in the readout box is bad - adds DC voltage to the signal. I stopped using GUR2_XYZ channels of the seism readout box. Now the same thing happened with the GUR1_XYZ channels. I checked the signals coming out from the seism box with the oscilloscope and they were fine. So the problem is not in the readout box. Then I applied 1 V sine wave to the input of AA board to the GUR1_X and ACC_MC1_Z channels. GUR1_X channel still shows noise. Something is wrong with these channels inside the AA board or in the ADC.

pem.png

 

Edited by Den: GUR1_XYZ_IN1 signals are empty though GUR1_XYZ are fine. So the problem is just that GUR1_XYZ_IN1 are not acquired for now though some of the ACC_IN1 channels contain the signal. I need to correct .ini files.

  6580   Fri Apr 27 12:12:14 2012 DenUpdateCDSc1ioo is back

Rolf came to the 40m today and managed to figure out what the problem is. Reading just dmesg was not enough to solve the problem. Useful log was in

>> cat /opt/rtcds/caltech/c1/target/c1x03/c1x03epics/iocC1.log

Starting iocInit
The CA server's beacon address list was empty after initialization?
iocRun: All initialization complete
sh: iniChk.pl: command not found
Failed to load DAQ configuration file

iniChk.pl checks the .ini file of the model.

>> cat /opt/rtcds/rtscore/release/src/drv/param.c


int loadDaqConfigFile(DAQ_INFO_BLOCK *info, char *site, char *ifo, char *sys)
{

  strcpy(perlCommand, "iniChk.pl ");
  .........
  strcat(perlCommand, fname); // fname - name of the .ini file
  ..........
}

So the problem was not in the C1X03.ini. The code could not find the perl script though it was in the /opt/rtcds/caltech/c1/scripts directory. Some environment variables are not set. Rolf added /opt/rtcds/caltech/c1/scripts/ to $PATH variable and c1ioo models (x03, ioo, gcv) started successfully. He is not sure whether this is a right way or not, because other machines also do not have "scripts" directory in their PATH variable.

>> cat /opt/rtcds/caltech/c1/target/c1x03/c1x03epics/iocC1.log

Starting iocInit
The CA server's beacon address list was empty after initialization?
iocRun: All initialization complete

Total count of 'acquire=0' is 2
Total count of 'acquire=1' is 0
Total count of 'acquire=0' and 'acquire=1' is 2

Counted 0 entries of datarate=256     for a total of 0
Counted 0 entries of datarate=512     for a total of 0
Counted 0 entries of datarate=1024     for a total of 0
Counted 0 entries of datarate=2048     for a total of 0
Counted 0 entries of datarate=4096     for a total of 0
Counted 0 entries of datarate=8192     for a total of 0
Counted 0 entries of datarate=16384     for a total of 0
Counted 0 entries of datarate=32768     for a total of 0
Counted 2 entries of datarate=65536     for a total of 131072

Total data rate is 524288 bytes - OK

Total error count is 0

Rolf mentioned about automatic set up of variables - kiis of smth like that - probably that script is not working correctly. Rolf will add this problem to his list.

  6579   Fri Apr 27 09:27:48 2012 DenUpdateCDSsus watchdogs?

Quote:

Why are all the suspension watchdogs tripped?  None of the suspension models are running on c1ioo, so they should be completely unaffected.  Steve, did you find them tripped, or did you shut them off?

In either event they should be safetly turned back on.

 I've turned off the coils. Though non of them are on the c1ioo, who knows what can happen when we'll try to run the models again.

  6578   Fri Apr 27 09:00:15 2012 JamieUpdateCDSsus watchdogs?

Why are all the suspension watchdogs tripped?  None of the suspension models are running on c1ioo, so they should be completely unaffected.  Steve, did you find them tripped, or did you shut them off?

In either event they should be safetly turned back on.

  6577   Fri Apr 27 08:11:19 2012 steveUpdateCDSc1ioo condition

 

 

Attachment 1: c1ioo.png
c1ioo.png
  6576   Thu Apr 26 20:44:23 2012 JamieUpdateCDSdaq network failure, c1ioo failing to start models

Den tried adding a SINGLE acquire channel to the c1ioo, which for some reason hung c1ioo and took down the entire DAQ network (at least all communication between the front ends and the fb).  We recovered by restarting c1ioo and restarting mx_stream on all the rest of the front-ends

After "recovering", though, c1ioo is failing to load models, or at least it's IOP. Here is the tail of dmesg when trying to start the IOP:

[ 1751.140283] c1x03: Initializing space for daqLib buffers
[ 1751.140284] c1x03: Initializing Network
[ 1751.140285] c1x03: Found 1 frameBuilders on network
[ 1751.250658] CPU 2 is now offline
[ 1751.250657] c1x03: Sync source = 4
[ 1751.250657] c1x03: Waiting for EPICS BURT Restore = 1
[ 1751.310008] c1x03: Waiting for EPICS BURT 0
[ 1751.310008] c1x03: BURT Restore Complete
[ 1751.310008] c1x03: Initialized servo control parameters.
[ 1751.311699] c1x03: DAQ Ex Min/Max = 1 3
[ 1751.311699] c1x03: DAQ XEx Min/Max = 3 53
[ 1751.311733] c1x03: DAQ Tp Min/Max = 10001 10007
[ 1751.311733] c1x03: DAQ XTp Min/Max = 10007 10507
[ 1751.311737] c1x03: DIRECT MEMORY MODE of size 64
[ 1751.311737] c1x03: daqLib DCU_ID = 33
[ 1751.311737] c1x03: Invalid num daq chans = 0
[ 1751.311737] c1x03: DAQ init failed -- exiting

The chan file for this model (/opt/rtcds/caltech/c1/chans/daq/C1X03.ini) looks totally fine, has two un-acquired channels uncommented, and has otherwise not been touched. The C1:FEC-33_MSGDAQ is also reading: "ERROR reading DAQ file!"

I'm at a loss for what is going on. I've tried restarting every CDS process on the machine, restarting the model multiple times, restarting fb, and even restarting the entire c1ioo machine, all to no affect.

  6575   Thu Apr 26 18:17:56 2012 SureshUpdateIOOMC WFS: Tweaked the WFS offsets

[Jamie, Suresh]

Yesterday Den and Koji reported that the WFS loops were causing the MC to become unlocked.  They had aligned the PMC.   The input beam into the MC seems to be well aligned.  MCREFL DC close to minimum it gets while MC is locked (~0.45 V).

I checked and saw that the WFS heads and the MC2_TRANS_QPD had picked up DC offsets.  To reset them I turned off the MC_autolocker and closed the PSL shutter.

The ADC offsets were set using this script /cvs/cds/rtcds/caltech/c1/scripts/MC/WFS/WFS_QPD_offsets.  (Jamie fixed the paths to ezcaservo to get this script to work)

The WFS sensor head offsets were manually set to adjust the Q and I signals from the sensor head to zero.  (This operation is supposed to be done by a script which is available, but I will check it out before I direct people to it).

Then we noticed that the ASC outputs were turned off.  (Presumably Koji turned them off yesterday, when the MC was repeatedly unlocking due to the WFS loops).

We turned on the ASC outputs and the MC stayed locked with reasonable outputs on the WFS output channels.  (+/-100)

However, engaging the WFS servo increases the MCREFL DC signal to 0.7 V from the 0.45 V value when the servo is not engaged.  This could be because of DC offsets in the WFS servo filters.   I will adjust these offsets to maintain good MC transmission when the WFS servo is engaged.

 

 

 

  6574   Thu Apr 26 18:15:59 2012 JamieUpdateCDSpossible issue with mx_stream on front ends

I'm noticing what appears to be occasional failures of mx_stream on the front end machines.  It doesn't happen that frequently, but I've noticed it a couple of times already since the upgrade.

The symptom is that the DC Status goes to "0xbad" (red) and the "FE NET" goes red for all models on a given front end.

The solution seems to be restarting mx_stream on the given front end:    sudo  /etc/init.d/mx_stream restart"

There is nothing in the mx_stream log:

 controls@c1sus ~ 0$ cat /opt/rtcds/caltech/c1/target/fb/mx_stream_logs/c1sus.log 
 c1x02
 c1sus
 c1mcs
 c1rfm
 c1pem
 mmapped address is 0x7f43740ec000
 mapped at 0x7f43740ec000
 mmapped address is 0x7f43700ec000
 mapped at 0x7f43700ec000
 mmapped address is 0x7f436c0ec000
 mapped at 0x7f436c0ec000
 mmapped address is 0x7f43680ec000
 mapped at 0x7f43680ec000
 mmapped address is 0x7f43640ec000
 mapped at 0x7f43640ec000
 send len = 263596
 Connection Made

but I do see some funny messages in the front end dmesg:

 [200341.317912] DXH Adapter 0 : Heartbeat alive-check for node=12 failed (cnt=8387 state=0x1 deb=0 val=0).
 [200341.318670] DXH Adapter 0 : Session for node 12 is disabled - Status = 0x5
 [200341.319062] Session callback reason=1 status=5 target_node=12
 [200341.319069] Session callback reason=3 status=0 target_node=12
 [200341.359534] (map_table_check_access:752):my id 1 ->  remote id 2 : entry was valid - is now tentatively valid
 [200341.859584] DXH Adapter 0 : Probe failure for node=12 - disabling session probeStatus=0x40000f02
 [200341.860335] DXH Adapter 0 : Session for node 12 is disabled - Status = 0x3
 [200341.860728] Session callback reason=1 status=3 target_node=12
 [200374.006111] DXH Adapter 0 : Set reachable remote node list.
 [200409.020670] DXH Adapter 0 : Set reachable remote node list.
 [200409.021076] DXH Adapter 0 : Session for node 12 is deleted - Status = 0x0
 [200409.021468] Session callback reason=5 status=0 target_node=12
 [200412.362824] (map_table_insert:648):** successfully inserted **(valid unicast) inst 0 node 1->0 fwd 0 fwd_tp 4 egress 0
 [200418.025994] (map_table_check_access:752):my id 1 ->  remote id 0 : entry was valid - is now invalid
 [200418.025998] (map_table_insert:648):** successfully inserted **(valid unicast) inst 0 node 1->2 fwd 0 fwd_tp 4 egress 0
 [200421.743916] Session callback reason=0 status=0 target_node=12
 [200422.073776] DXH Adapter 0 : Set reachable remote node list.
 [200422.342446] Session callback reason=7 status=0 target_node=12
 [200422.342454] DXH Adapter 0 : Session for node 12 is ok.

I'm awaiting feedback from experts.

 

  6573   Thu Apr 26 16:35:34 2012 JamieSummaryCDSrosalba now running Ubuntu 10.04

This morning I installed Ubuntu 10.04 on rosalba.  This is the same version that is running on pianosa.  The two machines should be identically configured, although rosalba may be missing some apt-getable packages.

  6572   Thu Apr 26 11:56:10 2012 DenUpdatePEMBlue Bird Pre-Amplifier

Quote:

Quote:

Usually R is for resistors and D is for diodes. Do you think from the schematic that we should put diodes into the R slots?

 That guys are resistors.

 You are right, they just looked like they were too small to be resistors when I glanced at them. 

  6571   Thu Apr 26 09:22:17 2012 steveUpdate under the east end optical table

Quote:

I added an U channel based bottom shelf at the south end today.

 I'm working on similar shelf at ETMY.  Precondition: NPRO in bypass mode, heater for doubling in bypass........since power outage?  Optical level servo turned off.......

U-channel based shelf in place.  Oplev servo is back on at 11:15am    The table may moved.  The oplev return is missing the quad by a few milimeter.

  6570   Wed Apr 25 21:24:10 2012 DenUpdateComputer Scripts / Programsc1oaf

C1OAF model, codes and medm screens are updated. All proper files are commited to svn and updated at the new model path.

  6569   Wed Apr 25 19:36:19 2012 DenUpdatePSLPMC aligned

[Koji, Den]

We have aligned PMC,  the WFS are not working yet.

  6568   Wed Apr 25 19:32:57 2012 DenUpdatePEMBlue Bird Pre-Amplifier

Quote:

Usually R is for resistors and D is for diodes. Do you think from the schematic that we should put diodes into the R slots?

 That guys are resistors.

  6567   Wed Apr 25 18:59:47 2012 ranaUpdatePEMBlue Bird Pre-Amplifier

Usually R is for resistors and D is for diodes. Do you think from the schematic that we should put diodes into the R slots?

  6566   Wed Apr 25 11:45:34 2012 JenneUpdatePEMBlue Bird Pre-Amplifier

It looks like we should change the "R40" and "R47" diodes.  If you do it this week, ask Jamie or Koji to check that you've got them oriented correctly before soldering them and plugging it in.

  6565   Wed Apr 25 00:20:01 2012 DenUpdatePEMacoustic noise at 40m

 Blue Bird Mic is suspended close to PMC now and outputs ~10 counts when pre-amp gain is 8 dB. This means that the mic outputs ~2.42 mV. Its sensitivity is 27 mV/Pa => acoustic noise is ~0.1 Pa or ~75 dB SPL.

If we buy Panasonic WM61A with their sensitivity -35 dB => they will output ~1.7 mV. We can amplify this signal without adding significant noise. For WM61A S/N ratio is given to be 62 dB. This is for some standard signal that is not specified. For Blue Bird mic it is specified according to IEC 651. So I assume SPL of the standard signal = 94 dB => noise level of WM61A is 32 dB (pretty bad compared to 7 dB-A of Blue Bird). But in our case for PSL S/N ratio is ~43 dB that is not too bad. PSL is noisy due to HEPA, acoustic noise level close to MC2 stack will be less. So we may want to consider Primo EM172/173 where the noise level is claimed to be 18 dB less. I think we should buy several WM61A and EM172.

  6564   Tue Apr 24 22:16:59 2012 DenUpdatePEMBlue Bird Pre-Amplifier

 I detached Clyde (pre-amp that was in the control room) to understand why it is not working. I seems that the board circuit is burnt near R40, R47, R45.

DSC_4273.JPG     DSC_4274.JPG

  6563   Tue Apr 24 16:15:24 2012 DenUpdatePEMmicrophones

I've installed Blue Bird microphone to listen to the acoustic noise at the PSL near PMC.

DSC_4271.JPG     DSC_4272.JPG

Coherence between MC_F and Blue Bird output (C1:PEM-ACC_MC2_Z for now) is changing from low to high value at frequencies 20 - 200 Hz with period ~1 min. Maybe HEPA works with some periodicity. Now it works pretty hard, ~80% of max.

micro_mc_low.png    micro_mc_high.png

  6562   Tue Apr 24 14:55:00 2012 JamieUpdateCDScds code paths (rtscore, userapps) have moved

NOTE: Unless you really care about what's going on under the hood, please ignore this entire post and go here: USE THE NEW RTCDS UTILITY

Quote:

  • the userapps directory is in a new place (/opt/rtcds/userapps).  Not everything in the old location was checked into the repository, so we need to check to make sure everything that needs to be is checked in, and that all the models are running the right code.

 An important new aspect of this upgrade is that we have some new directories and some of the code paths have moved to correspond with the "standard" LIGO CDS filesystem hierarchy:

  • rtscore:      /opt/rtcds/rtscore/release       -->  RTS RCG source code (svn: adcLigoRTS)
  • userapps: /opt/rtcds/userapps/release  -->  CDS userapps source for models, RCG C code, medm screens, scripts, etc (svn: cds_user_apps)
  • rtbuild:       /opt/rtcds/caltech/c1/rtbuild    -->  RCG build directory

All work should be done in the "userapps" directory, and all builds should be done in the build directory.  Some important points:

WARNING: DO NOT MODIFY ANYTHING IN RTSCORE

This is important.  The rtscore directory is now just where the RCG source code is stored.  WE NO LONGER BUILD MODELS IN THE RTS SOURCE  Please use the rtbuild directory instead.

NO MORE MODEL/CODE SYMLINKS

You don't need to link you model or code anywhere to get it to compile.  The RCG now uses proper search paths to source in the RCG_LIB_PATH to find the needed source.  This has been configured by the admin, so as long as you put your code in the right place it should be found.

ALL CODE/MODELS/ETC GO IN USERAPPS

All RTS code is now stored ENTIRELY in the userapps directory (e.g. /opt/rtcds/userapps/release/isc/c1/models/c1lsc.mdl).  This is more-or-less the same as before, except that symlinking is no longer necessary.  I have placed a symlink at the old location for convenience.

BUILD ALL MODELS IN RTSCORE

You can run "make c1lsc" in the rtbuild directory, just as you used to in the old core directory.  However, don't do that.  USE THE NEW RTCDS UTILITY instead.

  6561   Tue Apr 24 14:35:37 2012 JamieUpdateCDSlimited second trend lookback

Quote:

Alex told me that the "trend data is not available" message comes from the "trender" functionality not being enabled in daqd.  After re-enabling it (see #6555) minute trend data was available again.  However, there still seems to be an issue with second trends.  When I try to retrieve second trend data from dataviewer for which minute trend data *is* available I get the following error message:

Connecting to NDS Server fb (TCP port 8088)
Connecting.... done
No data found

read(); errno=9
read(); errno=9
T0=12-04-04-02-14-29; Length=3600 (s)
No data output.

Awaiting more help from Alex...

It looks like this is actually just a limit of how long we're saving the second trends, which is just not that long.  I'll look into extending the second trend look-back.

  6560   Tue Apr 24 14:30:08 2012 JamieUpdateCDSIntroducing: rtcds, a utility for interacting with the CDS RTS/RCG

The new rtcds utility

I have written a new utility for interacting with the CDS RTS/RCG system: "rtcds".  It should be available on all workstations and front-end machines, but certain commands are restricted to run on certain front end machines (build, start, stop, etc.).  Here's the help:

controls@c1lsc ~ 0$ rtcds help
Usage: rtcds <command> [arg]

Available commands:

  build|make SYS      build model
  install SYS         install model
  start SYS|all       start model
  restart SYS|all     restart running model
  stop|kill SYS|all   stop running model
  list                list all models for current host

controls@c1lsc ~ 0$ 

Please use this new utility from now on when interacting with RTS.

I'll be improving and expanding it as we go.   Please let me know if you encounter any problems.

 

  6559   Tue Apr 24 11:27:51 2012 steveUpdatePEMnew chairs

We received 6 new  chairs stools for the work benches that have 36" heights.

The control room east side and general office desks require 18" seat height. The new chairs stools have  minimum seat heights  26"

Please label chairs that we are getting rid of.

P4241026.JPG

  6558   Tue Apr 24 09:58:37 2012 steveUpdatePEMearthquake 3.9 magnitude

Local eq shakes the lab

Attachment 1: eq3.9.png
eq3.9.png
  6557   Mon Apr 23 23:20:07 2012 DenUpdatePEMmicrophones

Tonight I wanted to measure the ambient noise level using Blue Bird mics and figure out if Panasonic WM61a or Primo EM172/173 will be good enough or not. Blue Bird that is in the control room does not seem to work. May be the problem is with the pre-amplifier. The output measured by ADC/Oscilloscope is noise ( amplitude=5mV ). I will return to this issue tomorrow.

  6556   Sat Apr 21 21:10:34 2012 JamieUpdateCDStrend frame issue partially resolved

Quote:

Unfortunately it looks like there may be a problem with trend data, though.  If I try to retrieve 1 minute of "full data" with dataviewer for channel C1:SUS-ITMX_SUSPOS_IN1_DQ around GPS 1019089138 everything works fine:

Connecting to NDS Server fb (TCP port 8088)
Connecting.... done
T0=12-04-01-00-17-45; Length=60 (s)
60 seconds of data displayed

but if I specify any trend data (second, minute, etc.) I get the following:

Connecting to NDS Server fb (TCP port 8088)
Connecting.... done
Server error 18: trend data is not available
datasrv: DataWriteTrend failed in daq_send().
T0=12-04-01-00-17-45; Length=60 (s)
No data output.

Alex warned me that this might have happened when I was trying to test the new daqd without first turning off frame writing.

Alex told me that the "trend data is not available" message comes from the "trender" functionality not being enabled in daqd.  After re-enabling it (see #6555) minute trend data was available again.  However, there still seems to be an issue with second trends.  When I try to retrieve second trend data from dataviewer for which minute trend data *is* available I get the following error message:

Connecting to NDS Server fb (TCP port 8088)
Connecting.... done
No data found

read(); errno=9
read(); errno=9
T0=12-04-04-02-14-29; Length=3600 (s)
No data output.

Awaiting more help from Alex...

  6555   Sat Apr 21 20:40:28 2012 JamieUpdateCDSframebuilder frame writing working again

Quote:

A big one is that the framebuilder daqd started going haywire when I told it to start writing frames.  After restart the logs started showing this:

[Fri Apr 20 17:23:40 2012] main profiler warning: 0 empty blocks in the buffer
[Fri Apr 20 17:23:41 2012] main profiler warning: 0 empty blocks in the buffer
[Fri Apr 20 17:23:42 2012] main profiler warning: 0 empty blocks in the buffer
[Fri Apr 20 17:23:43 2012] main profiler warning: 0 empty blocks in the buffer
[Fri Apr 20 17:23:44 2012] main profiler warning: 0 empty blocks in the buffer
[Fri Apr 20 17:23:45 2012] main profiler warning: 0 empty blocks in the buffer
GPS time jumped from 1019002442 to 1019003041
FATAL: exception not rethrown
FATAL: exception not rethrown
FATAL: exception not rethrown

and the network seemed like it started to get really slow.  I wasn't able to figure out what was going on, so I shut the frame writing off again.  I'll have to work with Rolf on that next week.

So the framebuilder/daqd frame writing issue seems to have been a transient one.  Alex said he tried enabling frame writing manually and it worked fine, so I tried re-enabling the frame writing config lines and sure enough it worked fine.  So it's a mystery.  Alex said the "main profiler warning" lines tend to show up when data is backed up in the buffer, although he didn't explain why exactly that would have been the issue here.

daqdrc frame writing directives:

start frame-saver;
sync frame-saver;
start trender;
start trend-frame-saver;
sync trend-frame-saver;
start minute-trend-frame-saver;
sync minute-trend-frame-saver;
start raw_minute_trend_saver;