  2513   Wed Jan 13 12:03:00 2010 AlbertoUpdateABSLMeasurement now running. Please be careful

At the moment I'm running a measurement on the PRC and I'm planning to leave it going for the time we'll be at the 40m meeting.

Please be careful in the lab. Thank you.

  2521   Mon Jan 18 18:34:01 2010 AlbertoUpdateABSLMeasurement in progress

I started a long measurement of the PRC's transmissivity. I'm leaving the lab and I'm going to be back at about 8 tonight. Please do not disturb the interferometer. it is important that the MC and the PRC stay locked all the time.

  2522   Mon Jan 18 20:58:40 2010 AlbertoUpdateABSLMeasurement in progress

Quote:

I started a long measurement of the PRC's transmissivity. I'm leaving the lab and I'm going to be back at about 8 tonight. Please do not disturb the interferometer. it is important that the MC and the PRC stay locked all the time.

 That measurement is finished. I'm now going to start another one that will take another hour or so. I'm leaving it running for the night. If you want to work on the IFO, it should be definitely done by 11pm.

  2531   Tue Jan 19 12:54:39 2010 AlbertoUpdateABSLMeasurement in progress

A measurement will be running for the next hour. Please be careful.

  8591   Thu May 16 11:50:25 2013 KojiConfigurationElectronicsMeasurement and empirical models of the AI board TFs

Yesterday, I pulled out the AI board for the PRM/BS SUSs. (After the investigation it was restored)

Contrary to our expectation, the board D000186 was not Rev. A but Rev. B.

According to Jay's note in D000186 (for Rev.D), the differences of the Revs are as follows

Rev.A: Initial Release (Analog Biquad version, 4dB 4th order elliptic with notches)
Rev.B: Filter implemented by Freq Devices chip
Rev.C: Differential input version with better RF filtering
Rev.D: 3rd order 0.5dB ripple Cheby with notches at 16K&32K, DB25 input version


I went to the WB EE shop and found bunch of AI filter modules. At least I found one Rev.A and six Rev.D.
I found at least one Rev. C.

I took Rev.A and Rev. D to see the difference of the transfer functions.
Rev.A has more ripple but steeper roll-off. Rev. D is flater at the pass band with slower roll-off.
Rev.D has more phase lag, but it will be fine once the entire frequency response is shifted to x4 high frequency.
The notch frequency of the Rev. D looked right.

I made the empirical pole/zero modeling of the transfer functions.
The LISO models are attached as the ZIP file.
I faced an unexplainable phase behavior at around one the notches for Rev.A.
This may suggest there could have been internal saturation is the stage during the sweep.

More importantly, Rev. D has differential inputs although the connector formfactor is different from the current 40pin IDC.
In fact we should not use Rev.A or Rev.B as they have single end inputs.
Currently the inputs of the AI's for the SUSs are single ended while the DACs are differential.
This means that
1) We waste a half of the DAC range.
2) The negative outputs of the DACs are short-circuited. OMG
3) The ground level fluctuation between the DAC and the SUS rack fluctuates the actual actuation voltage.

Now I am looking at the noise performance of the filters as well as the DAC output noise and range.
I hope we can use Rev.D by replacing the connector heads as this will remove many of the problems we currently have.

Attachment 1: D000186AD_TF.pdf
D000186AD_TF.pdf
Attachment 2: D000186AD_TF.zip
  17035   Mon Jul 25 18:22:30 2022 DeekshaSummaryWikiMeasured the PZT TF Successfully

Measured the PZT beatnote using the setup mentioned in elog post 17031. Attached is the data taken from 10kHz to 1MHz, decadewise data was also taken that I'm not including in this post. A_R refers to the transfer function taken of channel A wrt the voltage reference (the swept sine we are inputting which has an IF of 30kHz). A and B correspond to the I and Q components of the signal taken from the DFD, respectively. I am currently working on plotting the data, and will shortly update this post with plots. Next steps - 

- quantify the uncertainty in the signal (I think)

- vectfit the data to find poles and zeroes

(and possibly find a better way to print/obtain data)

Edit: first pass of data plotted

Attachment 1: A_R_MAG.txt
"4395A REV1.12"
"DATE: Sep 17 2017"



"CHANNEL: 1"
"MEASURE TYPE: A/R"
"FORMAT TYPE: LOG MAG"
"NUMBER of POINTS: 801"
"SWEEP TIME:  385.3 ms"
... 811 more lines ...
Attachment 2: A_R_PHASE.txt




"CHANNEL: 2"
"MEASURE TYPE: A/R"
"FORMAT TYPE: PHASE (DEG)"
"NUMBER of POINTS: 801"
"SWEEP TIME:  385.3 ms"
... 808 more lines ...
Attachment 3: B_R_MAG.txt
"4395A REV1.12"
"DATE: Sep 17 2017"



"CHANNEL: 1"
"MEASURE TYPE: B/R"
"FORMAT TYPE: LOG MAG"
"NUMBER of POINTS: 801"
"SWEEP TIME:  385.3 ms"
... 809 more lines ...
Attachment 4: B_R_PHASE.txt




"CHANNEL: 2"
"MEASURE TYPE: B/R"
"FORMAT TYPE: PHASE (DEG)"
"NUMBER of POINTS: 801"
"SWEEP TIME:  385.3 ms"
... 807 more lines ...
Attachment 5: freq_resp_I.png
freq_resp_I.png
Attachment 6: freq_resp_Q.png
freq_resp_Q.png
  17603   Thu May 25 14:40:03 2023 PacoUpdateCalibrationMeasured the PSL wavelength

The PSL wavelength is 1064.5068 +- 0.0015 nm

[Paco, Yehonathan]

With yehonathan's help, we borrowed a WS6-200 wavemeter and calibrated the PSL wavelength using the frequency doubled beam at the PSL table. For this we gathered a FC/APC to FC/PC fiber, cleaned both ends, and launched ~ 6 uW of power into the wavemeter (Attachment #1 shows the setup). The vacuum wavelength was measured to be 532.2534 nm, (see picture in Attachment #2) implying a PSL wavelength of 1064.5068 nm. This is not too far from my previous estimate! The wavemeter claims a "200 MHz accuracy" so we used this as a standard error to estimate the uncertainty of  2 * 0.7 pm @ 532 nm = 1.5 pm @ 1064 nm. This leaves us with 1.4 ppm of relative wavelength error in the ALS based calibration.

  • It should be noted that the Nd:YAG temperature was set to 30.615(5) degrees (?) and the injection current to 2.100 Amps for this measurement.
Attachment 1: PXL_20230525_205542591~2.jpg
PXL_20230525_205542591~2.jpg
Attachment 2: PXL_20230525_205332570.jpg
PXL_20230525_205332570.jpg
  11550   Mon Aug 31 14:15:23 2015 IgnacioUpdateIOOMeasured the MC_F whitening poles/zeroes

I measured the 15 Hz zero and the 150 Hz pole for the whitening filter channels of the Generic Pentek board in the IOO rack. The table below gives these zero/pole pairs for each of the 8 channels of the board.

channel zero [Hz] pole [Hz] Chan
1 15.02 151.05 C1:ASC-POP_QPD_YAW
2 15.09 150.29 C1:ASC-POP_QPD_PIT
3 14.98 150.69 C1:ASC-POP_QPD_SUM
4 14.91 147.65 C1:ALS-TRX
5 15.03 151.19 C1:ALS-TRY
6 15.01 150.51 ---
7 14.95 150.50 C1:IOO-MC_L
8 15.03 150.93 C1:IOO-MC_F

Here is a plot of one of the measured transfer functions,

and the measured data is attached here: Data.zip


EQ: I've added the current channels going through this board. 

More importantly, I found that the jumpers on channel one (QPD X) were set to no whitening, in contrast to all other channels. Thus, the POP QPD YAW signals we've been using for who knows how long have been distorted by dewhitening. This has now been fixed. 

Hence, the current state of this board is that the first whitening stage is disabled for all channels and the second stage is engaged, with the above parameters. 

Attachment 1: Data.zip
  16964   Thu Jun 30 17:19:55 2022 DeekshaSummaryElectronicsMeasured Transfer Functions of the Control Loop, Servo (OLTF); got Vectfit working

[Cici, Deeksha]

We were able to greatly improve the quality of our readings by changing the parameters in the config file (particularly increasing the integration and settle cycles, as well as gradually increasing our excitation signals' amplitude). Attached are the readings taken from the same (the files directly printed by ssh'ing the SR785 (apologies)) - Attachment 1 depicts the graph w/ 30 data points and attachment 2 depicts the graph with 300 data points. 

Cici successfully vectfit to the data, as included in Attachment 3. (This is the vectfit of the entire control loop's OLTF). There are two main concerns that need to be looked into, firstly, the manner in which to get the poles and zeros to input into the vectfit program. Similarly, the program works best when the option to enforce stable poles is disabled, once again it may be worth looking into how the program works on a deeper level in order to understand how to proceed. 

Just as the servo's individual transfer function was taken, we also came up with a  plan to measure the PZT's individual transfer function (using the MokuLab). The connections for the same have been made and the Moku is at the Xend (disconnected). We may also have to build a highpass filter (similar to the one whose signal enters the PZT) to facilitate taking readings at high frequencies using the Moku. 

Attachment 1: TFSR785_29-06-2022_114042.pdf
TFSR785_29-06-2022_114042.pdf
Attachment 2: TFSR785_29-06-2022_114650.pdf
TFSR785_29-06-2022_114650.pdf
Attachment 3: TF_OLG_vectfit.png
TF_OLG_vectfit.png
  9661   Mon Feb 24 13:21:00 2014 JenneUpdateElectronicsMeasured REFL165 demod board

I measured the REFL 165 demod board's I/Q separation. 

Our 11MHz signal is currently 11.066092 MHz, so I put a signal to the RF input of the REFL165 demod board at 165.992380 MHz (15*11 MHz + 1kHz), with a signal of -13 dBm.

I then used the SR785 to measure the transfer function between the I and Q output channels. 

I got 82.7 degrees, at -0.64 dB. (I don't remember now if I had I/Q, or Q/I, not that it really matters). So, it seems that the REFL165 demod board has good separation, and at least isn't totally broken.

  9666   Mon Feb 24 17:59:31 2014 RANAUpdateElectronicsMeasured REFL165 demod board

 

 Demod boards should be at 90 deg, not 82.7 or 12 or yellow or ****. We should re-inject the RF and then set the D Phase in the filter module to make the signals orthogonal. 165 is a challenging one to get right, but its worth it since the signals are close to degenerate already.

  10967   Tue Feb 3 04:07:49 2015 JenneUpdateModern ControlMeasured PRM -> POP QPD TFs

Today (after centering the POP QPD), I measured the PRM to POP QPD transfer functions.  I am suspicious that this was part of my problem last week.  Since most of the angular noise is coming from the folding mirrors, but I can't actuate on them, I need to know (rather, the Wiener calculator needs to know) how actuating on the PRM will affect the spot at the POP QPD.

For the plots below, I have cut out any data points that have coherence less than 0.95. I may want to go back and fill in a little bit some of the areas (particularly around 3Hz) that I had trouble getting coherence in.

Using these to prefilter my witness data, I am failing to calculate a Wiener filter.  I have tried the Levinson algorithm, as well as brute-forcing it, but I'm too close to singular for either to work.  I am able to calculate a set of Wiener filters without the prefiltering, or with a dummy very simple prefilter, so it's not inherently in the calculators.  Separately, I can plot my vectfit-ed actuator TFs, and I can convert them to a discrete fiilter with the bilinear transform, and then use sosfilt to filter some white noise data, which comes out with the shape I expect.  So.  It's not inherently the filters either.  More work to do, when it's not 4am.

Here are the measured actuator transfer functions.  They were measured as usual with DTT, but since the measurement kept getting interrupted (MC unlock, or I wanted to add more integrations or more cycles), these are several different DTT files stitched together.  In both cases I am acuating PRM's ASC[pit, yaw] EXC point, and looking at the POP QPD.

 

Attachment 1: PIT_measured_TF.png
PIT_measured_TF.png
Attachment 2: YAW_measured_TF.png
YAW_measured_TF.png
  2332   Wed Nov 25 14:29:08 2009 robUpdateLockingMeasured MC length--FSS trend

Quote:

Quote:

What I meant was the VCO driver, not the FSS box.

As for the frequency, all written numbers were the Marconi displays.
The number on the frequency counter was also recorded, and so will be added to the previous entry shortly... 

Quote:

I propose that from now on, we indicate in the elog what frequencies we're referring to. In this case, I guess its the front panel readback and not the frequency counter -- what is the frequency counter readback? And is everything still locked to the 10 MHz from the GPS locked Rubidium clock?

Plus, what FSS Box? The TTFSS servo box? Or the VCO driver? As far as I know, the RC trans PD doesn't go through the FSS boxes, and so its a real change. I guess that a bad contact in the FSS could have made a huge locking offset.

 

 

Locking has gone sour.  The CARM to MCL handoff, which is fairly early in the full procedure and usally robust, is failing reliably. 

As soon as the SUS-MC2_MCL gain is reduced, lock is broken.  There appears to be an instability around 10Hz.  Not sure if it's related.

 Five day minute trend.  FAST_F doesn't appear to have gone crazy.

Attachment 1: FSStrendpowerjump.png
FSStrendpowerjump.png
  2335   Wed Nov 25 16:13:27 2009 ranaUpdatePSLMeasured MC length--FSS trend

but the increase in both the RCtrans and the RCrefl is consistent with my theory that the power going to the RC has increased ; its not just an increase in the visibility.

We should scan the AOM/VCO to make sure the frequency is matched to the resonance to within 0.5 dB.

  2336   Wed Nov 25 16:44:52 2009 KojiUpdatePSLMeasured MC length--FSS trend

I checked C1:PSL-FSS_VCODETPWR. The attached is the 4 months trend of the FSS RCTRANS / RFPDDC(=FSS REFL) / VCODETPWR / VCOMODLEVEL.

Although VCO modulation level setting was mostly constnt, VCODETPWR, which presumably represents the RF level, changes time by time.
It coincides with the recent reduction of the RCTRANS/RFPDDC. Actually, my touch restored the VCO to the previous more stable state.
One can see that this is not only a single occation, but it happened before too. (In the middle of Aug.)

This could be explained by the bad contact of some cable or connector.

Nevertheless we need more careful investigation:

1. Understand what VCODETPWR is exactly.
2. Investigate relationship between VCOMODLEVEL / VCODETPWR / AOM deflection efficiency / RCTRANSPD
3. Confirm the frequency matching between the VCO and AOM.

Quote:

but the increase in both the RCtrans and the RCrefl is consistent with my theory that the power going to the RC has increased ; its not just an increase in the visibility.

We should scan the AOM/VCO to make sure the frequency is matched to the resonance to within 0.5 dB.

 

Attachment 1: 091125_FSS.png
091125_FSS.png
  2317   Mon Nov 23 21:30:29 2009 JenneUpdateLSCMeasured MC length

With Koji's help, I measured the length of the Mode Cleaner.

The new modulation frequencies (as quoted on the Marconi front panels) are: 

165.980580 MHz

 33.196116 MHz

132.784464 MHz

199.176696 MHz

The Frequency Counter readback is 165980584.101 Hz (a 4Hz difference).  All of the Marconi's front-panel frequencies read ###.##### MHz Ext, and the Frequency standard has it's "locked" light illuminated, and the 1pps input light blinking, so I think everything is still nicely locked to the frequency standard, and the frequency standard is locked to the GPS.

While changing the marconi's, I accidentally touched the MC's 29.5 MHz marconi.  It is set back to the nominal value (according to Kiwamu's rack photos) of 29.485MHz.  But the phase might be sketchy, although hopefully this doesn't matter since we don't do a double demodulation with it.

I also ran the scripts in the wiki page: How To/Diagonalize DRMI Length Control to set the DD Phases.

 

 

  2319   Tue Nov 24 08:00:16 2009 ranaUpdateLSCMeasured MC length

I propose that from now on, we indicate in the elog what frequencies we're referring to. In this case, I guess its the front panel readback and not the frequency counter -- what is the frequency counter readback? And is everything still locked to the 10 MHz from the GPS locked Rubidium clock?

Plus, what FSS Box? The TTFSS servo box? Or the VCO driver? As far as I know, the RC trans PD doesn't go through the FSS boxes, and so its a real change. I guess that a bad contact in the FSS could have made a huge locking offset.

 

  2320   Tue Nov 24 10:36:21 2009 KojiUpdateLSCMeasured MC length

What I meant was the VCO driver, not the FSS box.

As for the frequency, all written numbers were the Marconi displays.
The number on the frequency counter was also recorded, and so will be added to the previous entry shortly... 

Quote:

I propose that from now on, we indicate in the elog what frequencies we're referring to. In this case, I guess its the front panel readback and not the frequency counter -- what is the frequency counter readback? And is everything still locked to the 10 MHz from the GPS locked Rubidium clock?

Plus, what FSS Box? The TTFSS servo box? Or the VCO driver? As far as I know, the RC trans PD doesn't go through the FSS boxes, and so its a real change. I guess that a bad contact in the FSS could have made a huge locking offset.

 

  2325   Wed Nov 25 03:05:15 2009 robUpdateLockingMeasured MC length

Quote:

What I meant was the VCO driver, not the FSS box.

As for the frequency, all written numbers were the Marconi displays.
The number on the frequency counter was also recorded, and so will be added to the previous entry shortly... 

Quote:

I propose that from now on, we indicate in the elog what frequencies we're referring to. In this case, I guess its the front panel readback and not the frequency counter -- what is the frequency counter readback? And is everything still locked to the 10 MHz from the GPS locked Rubidium clock?

Plus, what FSS Box? The TTFSS servo box? Or the VCO driver? As far as I know, the RC trans PD doesn't go through the FSS boxes, and so its a real change. I guess that a bad contact in the FSS could have made a huge locking offset.

 

 

Locking has gone sour.  The CARM to MCL handoff, which is fairly early in the full procedure and usally robust, is failing reliably. 

As soon as the SUS-MC2_MCL gain is reduced, lock is broken.  There appears to be an instability around 10Hz.  Not sure if it's related.

  2333   Wed Nov 25 15:38:08 2009 robUpdateLockingMeasured MC length

Quote:

Quote:

What I meant was the VCO driver, not the FSS box.

As for the frequency, all written numbers were the Marconi displays.
The number on the frequency counter was also recorded, and so will be added to the previous entry shortly... 

Quote:

I propose that from now on, we indicate in the elog what frequencies we're referring to. In this case, I guess its the front panel readback and not the frequency counter -- what is the frequency counter readback? And is everything still locked to the 10 MHz from the GPS locked Rubidium clock?

Plus, what FSS Box? The TTFSS servo box? Or the VCO driver? As far as I know, the RC trans PD doesn't go through the FSS boxes, and so its a real change. I guess that a bad contact in the FSS could have made a huge locking offset.

 

 

Locking has gone sour.  The CARM to MCL handoff, which is fairly early in the full procedure and usally robust, is failing reliably. 

As soon as the SUS-MC2_MCL gain is reduced, lock is broken.  There appears to be an instability around 10Hz.  Not sure if it's related.

 Whatever the locking problem was, the power of magical thinking has forced it to retreat for now.  The IFO is currently locked, having completed the full up script.  One more thing for which to be thankful.

  1662   Tue Jun 9 11:29:07 2009 JenneUpdateoplevsMeasured ETMY oplev beam size...put everything back

I measured the ETMY oplev beam size at a couple different distances away from the HeNe by taking out the steering mirror and letting the light propagate a ways.  I put the steering mirror back, aligned the oplev, and was able to relock the Yarm, so I think it's all back as it has been the last couple of weeks.

 

Now I need t o do some geometry and ray-tracing matrices to decide what focal length lens to buy, then we'll have  a shiny new ETMY oplev. 

  13080   Mon Jun 26 09:39:15 2017 NaomiSummaryGeneralMeasure transfer functions of Mini-Circuits filters

I have spent my first few days as a SURF getting experience working with the Network/Spectrum Analyzer (AG 4395A). After an introduction to the 40m by Koji, I was tasked with using the AG4395A to measure the transfer function of several filters (for example, Mini-Circuits Low Pass Filter SLP-30). I am now familiar with configuring the AG 4395A, taking a single set of data using a command from one of the control computers, and plotting the dataset as a Bode plot (separate plots for magnitude and phase) using Python.

To Do:

  • Use AGmeasure to take multiple datasets with a single command.
  • Plot multiple datasets for each filter on a single Bode plot and perform some statistical analysis. 

To experiment with plotting multiple datasets on a single Bode plot, I used a single dataset from the Network Analyzer using the SLP-30 filter and added random noise to create ten datasets to plot. I am attaching the resulting Bode plot, which has the ten generated sets of data plotted along with their average.

We discussed with Rana and Koji how to interpret this type of dataset from the Network Analyzer. Instead of considering the magnitude and phase as separate quantities, we should consider them together as a single complex number in the form H(f) = M exp(iπP/180), where M is the magnitude and P is the phase in degrees. We can then find the average value of the measured quantity in its complex number form (x + iy), as opposed to just taking the average of the magnitude and phase separately.  

Attachment 1: Bode_Plot.png
Bode_Plot.png
  545   Thu Jun 19 15:52:06 2008 AlbertoConfigurationComputersMeasure of the current absorbed by the new Megatron Computer
Together with Rich Abbot, sam Abbot and I measured the current absorbed by the new Megatron computer that we installed yesterday in the 1Y3 rack. The computer alone absorbs 8.1A at the startup and then goes down to 5.9A at regime. The rest of the rack took 5.2A without the computer so the all rack needs 13.3 at the startup and the 11.1A.

We also measured the current for the 1Y6 rack where an other similar Sun machine has been installed as temporary frame builder and we get 6.5A.


Alberto, Rich and Sam Abbot
  1503   Mon Apr 20 20:00:44 2009 ranaConfigurationIOOMcWFS gains re-allocated
Since it looks like the night time people have been running with a WFS gain of 0.05 and I like the slider
to be at 1.0, I lowered all of the WFS1/2_P/Y gains by 10 and increased the overall slider from 0.05 to 1.0.
So the loop gains are now 2x higher; with it like this I guess the UGFs are in the ~0.2-0.5 Hz range.
  439   Tue Apr 22 22:51:30 2008 ranaConfigurationIOOMcWFS Status
I've been working a little on the MC WFS in the last few days. I have made many
changes to the sensing matrix script and also to the MCWFSanalyze.m script.

The output matrix, as it was, was not bad at low frequencies but was making noise in
the ~1 Hz band. Turning the gain way down made it do good things at DC and not make
things work higher.

The output matrix generating script now works after Rob fixed the XYCOM issue. Not sure
what was up there. As Caryn mentioned the SUS2.ini channels were all zero after Andrey's
PEM power cycle a few days ago. Rob booted c1susvme to get the SUS1 channels back and
today we did c1susvme2 to get the IOO-MC_L et. al. back.

Even after doing the matrix inversion there is some bad stuff in the output matrix. I
checked that the sensing matrix measurement has good coherence and I measured and set the
MC WFS RF phases (they were off by ~20-30 deg.). Still no luck.

My best guess now is that the RG filters I've used for POS damping and the movement of the
beam on the MC mirror faces has made a POS<->YAW instability at low frequencies. My next
move is to revert to velocity damping and see if things get better. Should also try redoing
the A2L on the MC1-3.
  15988   Thu Apr 1 21:13:54 2021 AnchalUpdateSUSMatrix results, new measurement set to trigger
New Input matrix used for MC2 (C1:SUS-MC2_INMATRIX_ii_jj
  UL UR LR LL SIDE
POS 0.2464 0.2591 0.2676 0.2548 -0.1312
PIT 1.7342 0.7594 -2.494 -1.5192 -0.0905
YAW 1.2672 -2.0309 -0.9625 2.3356 -0.2926
SIDE 0.1243 -0.1512 -0.1691 0.1064 0.9962

New output matrix for MC2 (C1:SUS-MC2_TO_COIL_ii_jj_GAIN)
  POS PIT YAW
UL 1 1.022 0.6554
UR 1 0.9776 -1.2532
LL 1 -0.9775 1.2532
LR 1 -1.0219 -0.6554

Measured Sensing Matrix (Cross Coupling) (Sensed DOF x Excited DOF)
  Excited POS Excited PIT Excited YAW
Sensed POS 1 1.9750e-5 -3.5615e-6
Sensed PIT 0 1 -6.93550e-2
Sensed YAW 0 -2.4429e-4 1

A longer measurement is set to trigger at 5:00 tomorrow on April 2nd, 2021. This measurement will run for 35 iterations with an excitation duration of 120s and bandwidth for CSD measurement set to 0.1 Hz. The script is set to trigger in a tmux session named 'cB' on pianosa.

  15993   Fri Apr 2 15:22:54 2021 gautamUpdateSUSMatrix results, new measurement set to trigger

How should I try to understand why PIT and YAW are so different? 

Quote:
New output matrix for MC2 (C1:SUS-MC2_TO_COIL_ii_jj_GAIN)
  POS PIT YAW
UL 1 1.022 0.6554
UR 1 0.9776 -1.2532
LL 1 -0.9775 1.2532
LR 1 -1.0219 -0.6554
  8774   Thu Jun 27 21:59:42 2013 ranaUpdateComputer Scripts / ProgramsMatlab upgraded

I moved the old matlab directory from /cvs/cds/caltech/apps/linux64/matlab_o to /cvs/cds/caltech/apps/linux64/matlab_oo

and moved the previously current matlab dir from /cvs/cds/caltech/apps/linux64/matlab to /cvs/cds/caltech/apps/linux64/matlab_o.

And have installed the new Matlab 2013a into /cvs/cds/caltech/apps/linux64/matlab.

Since I'm not sure how well the new Matlab/Simulink plays with the CDS RCG, I've left the old one and we can easily revert by renaming directories.

  777   Thu Jul 31 16:11:22 2008 josephbConfigurationComputersMatlab on Megatron
Matlab now works on megatron.

I did a few things:

1) Added to the PATH environment variable. Did this in .bash_profile in the /home/controls directory by adding the line

PATH=$PATH:/cvs/cds/caltech/apps/linux64/matlab/bin/
export PATH

This probably should be somewhere else up further up the line, but I was too lazy to figure it out.

2)Fixed a gateway mistake I had added earlier so the megatron could use the NAT router and see the outside world so yum worked.

3) Removed the i386 based libXp and openmotif packages.

4) Installed the x86_64 based libXp and openmotif packages.

Edit: Forgot that I also added the following line to the /etc/fstab file in order to mount the shared code. This was stolen directly from Rosalba's /etc/fstab file. This was so that it could see the matlab code.
linux1:/home/cds/ /cvs/cds nfs rw,bg,soft 0 0
  10751   Wed Dec 3 21:41:12 2014 JenneUpdateComputer Scripts / ProgramsMatlab license updated

It seems that the old Matlab servers went down a week or so early, so I have updated the Matlab license information in 

/cvs/cds/caltech/apps/linux64/matlab/licenses/

per the instructions on https://www.imss.caltech.edu/content/updating-matlab-license-file

EDIT: Q did this also for the control room iMac

  281   Mon Jan 28 17:16:54 2008 AndreyConfigurationComputersMatlab libraries DO NOT WORK properly sometimes

Working in Matlab, I encountered at two different times today the license distribution problem:

??? License checkout failed.
License Manager Error -4
Maximum number of users for Curve_Fitting_Toolbox reached.
Try again later.
To see a list of current users use the lmstat utility or contact your License Administrator.

Troubleshoot this issue by visiting:
http://www.mathworks.com/support/lme4a
  6083   Wed Dec 7 20:55:44 2011 Vladimir, DenUpdatedigital noiseMatlab error

Quote:

It would be useful to see some plots so we could figure out exactly what magnitude and phase error correspond to "gross" and "miserable".

To show why Matlab is bad in filtering at small cut-off frequencies we did the same thing in Matab, Octave and R: we've taken the low-pass chebyshev filter of the type 1, order 6, ripple 1 dB, the sampling frequency was 16384 Hz and cut-off frequency varied from 1 to 1000 Hz. Here is the plot for the gain of the zpk model versus to cut-off frequency.

gain_cmp.png

We can see that Matlab's gain shows surprising gains for low cut-off frequencies through for > 100 Hz it is fine. In the next table we compare gain from Foton, Matlab, R and Octave for the same filter. So Foton is also good

freq R_gain matlab_gain octave_gain Foton_gain
1 3.05186270655452e-24 4.8824152e-22 3.05186271e-24 3.05186e-24
10 3.04698335947093e-18 1.8268825e-16 3.04698336e-18 3.04698e-18
100 2.99910841393644e-12 2.9991084e-12 2.99910841e-12 2.99911e-12
1000 2.60247212721439e-06 2.6024721e-06 2.60247213e-06 2.60247e-06
  12859   Wed Mar 1 16:00:41 2017 gautamUpdateComputer Scripts / ProgramsMatlab R2016b installed

Since it would be nice to have the latest version of Matlab, with all its swanky new features (?), available on the control room computers and Optimus, I downloaded Matlab R2016b and activated it with the Caltech Campus license. I installed it into /cvs/cds/caltech/apps/linux64/matlab16b. Specifically, I would like to run the coating optimization code on Optimus, where I can try giving it more stringent convergence criterion to see if it converges to a better spot.

I trust that this way, we don't interfere with any of the rtcds stuff.

If I've done something illegal license-wise or if this is likely to cause havoc, please point me to what is the correct way to do this.

GV 18 Mar 2017: Though I installed this using the campus network license key, this seems to only work on Rossa. If I run it on the other control room machines/Optimus, it throws up a licensing error. I will check with Larry W. as to how to resolve this...

 

  251   Mon Jan 21 23:30:03 2008 AndreyUpdateComputer Scripts / ProgramsMatlab Program for Q-factor measurements (XARM -> ITMX and ETMX)

Finally I overcame difficulties with adapting Sonia's Matlab programs for XARM (Sonia's program was for MC),

and now there exists a Matlab program that makes a fit of a ringdown curve and calculates Q-factor for a mirror ITMX.

Specifically, this program allows to measure ringdown, fit it and calculate Q-factor for the ITMX-mirror for a specific value of
"C1:SUS-ITMX_SUSPOS_GAIN".

Attached is a plot of a ringdown curve and its fit for the value 4.0 in channel "C1:SUS-ITMX_SUSPOS_GAIN".

Calculations yield the result Q=3.7+-0.2 for the value 4.0 in channel "C1:SUS-ITMX_SUSPOS_GAIN".

As Robert started 10 minutes ago the long procedure of the whole interferometer locking,
I cannot disturb the interferometer now, so I will measure Q-factors for various combinations of suspension damping gain on Tuesday.

I will also easily modify the program for measuring Q-factors of ETMX-mirror and make measurements with ETMX on Tuesday.

The Matlab scripts are in directory /cvs/cds/caltech/users/rodionov/Q-Factors/
Attachment 1: Example-ITMX_POS_40.png
Example-ITMX_POS_40.png
  120   Tue Nov 20 18:35:20 2007 JohnHowToComputersMatLab in Emacs
If you can't get MatLab to run in emacs try adding the following to the .emacs file

(setq matlab-shell-command-switches '("-nojvm"))

This stops the gui opening.

To start MatLab type M-x matlab-shell.
To enter MatLab mode M-x matlab-mode.

I've done this on LINUX3.

To run MatLab in emacs under windows one can use MatLabShell http://www.cs.umb.edu/~ram/matlabShell/index.html
  3740   Tue Oct 19 00:24:07 2010 DmassOmnistructureElectronicsMassive restocking of the 40m

I had a number of delinquent items on the sign out list from the 40m. I returned about half, and ordered replacements for most of the other half.

I put the photodiodes on the SP table, and the 560 on the electronics  bench.

  17612   Fri Jun 2 10:59:20 2023 advaitUpdatePEMMass for temperature control

I measured the dimensions of the puck we shall be using as the mass for the toy model. It has a diameter of 75 mm and a thickness of 25 mm. Assuming a density of 2.7 g/cm^3, this brings the mass to 298.2 g.

  17613   Fri Jun 2 11:47:33 2023 KojiUpdatePEMMass for temperature control

How about measuring the actual weight with a scale? There are a couple of scales on Yuta's desk.

  17614   Fri Jun 2 16:45:46 2023 advaitUpdatePEMMass for temperature control

Thanks, I was actually looking for a scale earlier but could not find it after asking a couple of people. One of the scales reads 298 g, while the other has a limit of 200g. Looks like it is indeed aluminium.

Quote:

How about measuring the actual weight with a scale? There are a couple of scales on Yuta's desk.

 

  16135   Wed May 12 14:23:20 2021 JordanUpdateSUSMass Properties of SOS Assembly with 3"->2" Optic sleeve, in SI units
Attachment 1: Moments_of_Inertia_SI.PNG
Moments_of_Inertia_SI.PNG
  16136   Wed May 12 16:53:59 2021 KojiUpdateSUSMass Properties of SOS Assembly with 3"->2" Optic sleeve, in SI units

No, this is the property of the suspension assembly. The mass says 10kg

Could you do the same for the testmass assembly (only the suspended part)? The units are good, but I expect that the values will be small. I want to keep at least three significant digits.

  16137   Wed May 12 17:06:52 2021 JordanUpdateSUSMass Properties of SOS Assembly with 3"->2" Optic sleeve, in SI units

Here are the mass properties for the only the test mass assembly (optic, 3" ring, and wire block). (Updated with g*mm^2)

Quote:

No, this is the property of the suspension assembly. The mass says 10kg

Could you do the same for the testmass assembly (only the suspended part)? The units are good, but I expect that the values will be small. I want to keep at least three significant digits.

 

Attachment 1: Moments_of_Inertia_SI.PNG
Moments_of_Inertia_SI.PNG
  16146   Wed May 19 18:29:41 2021 KojiUpdateSUSMass Properties of SOS Assembly with 3"->2" Optic sleeve, in SI units

Calculation for the SOS POS/PIT/YAW resonant frequencies

- Nominal height gap between the CoM and the wire clamping point is 0.9mm (cf T970135)

- To have the similar res freq for the optic with the 3" metal sleeve is 1.0~1.1mm.
As the previous elog does not specify this number for the current configuration, we need to asses this value and the make the adjustment of the CoM height.

Attachment 1: SOS_resonant_freq.pdf
SOS_resonant_freq.pdf SOS_resonant_freq.pdf
Attachment 2: SOS_resonant_freq.nb.zip
  9026   Mon Aug 19 09:54:13 2013 SteveUpdatesafetyMasayuki receives safety training

Masayuki Nakano, a student of Seiji's from ICRR / U Tokyo, is visiting us here at the 40m lab for the next couple months.

He received 40m specific basic safety training this morning.

  1325   Thu Feb 19 16:29:43 2009 YoichiUpdateComputersMartian wireless router bad
The Martian wireless router is dead.
I rebooted it several times, but it hangs up in a minute.
I will ask steve to buy a new one.
  1339   Thu Feb 26 01:24:44 2009 YoichiUpdateComputersMartian wireless is back
Today, a new wireless router arrived.
I configured and installed it. Now the martian wireless network is back.
I updated the wiki page about the wireless network.
http://lhocds.ligo-wa.caltech.edu:8000/40m/Network
  10196   Mon Jul 14 16:51:07 2014 NichinUpdateElectronicsMartian table updated, Named server restarted

 [Nichin, Jenne]

The martian lookup tables are located at /etc/bind/zones/martian.db  and etc/bind/zones/rev.113.168.192.in-addr.arpa

Jenne updated these to include santuzza.martian  192.168.113.109

 

 

The method to restart named server given at  https://wiki-40m.ligo.caltech.edu/Martian_Host_Table  also does not work.

I restarted it using  >sudo /etc/init.d/bind9 restart

The named server is now updated and works fine. :)  I will update the 40m wiki now.

  14465   Tue Feb 19 19:03:18 2019 ranaUpdateComputersMartian router -> WPA2

I have swapped our martian router's WiFi security over to WPA2 (AES) from the previous, less-secure, system. Creds are in the secrets-40-red.

  4135   Tue Jan 11 14:05:11 2011 josephbUpdateComputersMartian host table updated daily

I created two simple cron jobs, one running on linux1 and one running on nodus, to produce an updated copy of the martian host table linkable from the wiki every day.

The scripts live in /opt/rtcds/caltech/c1/scripts/AutoUpdate/.  One is called  updateHostTable.cron and run on linux1 everyday at 4 am, and the other is called moveHostTable.cron which is run on nodus everyday at 5am.

The new link has been added to the Martian Host table wiki page  here.

 

  4603   Tue May 3 00:44:02 2011 KojiConfigurationComputersMartian WIreless Bridge

The Martian wireless bridge has the ethernet cable inserted in the wrong connector.

It should be inserted to one of the four port. Not in the "INTERNET" connector.

Once the connector has been changed, the martian net as well as the internet became accessible from the laptops.

ELOG V3.1.3-