  4294   Tue Feb 15 02:13:16 2011 kiwamuSummaryASCa daytime task : small signals on ETMX OL

Rana and I found that the QPD for the optical lever at X end are showing small signals.

At this moment each of the segments exhibits approximately 200 counts when the oplev beam is centered.

These small numbers may be due to the coating of ETMX, but we are not sure.

Probably we have to increase the gain of the QPD depending on situations.

 

So a set of the tomorrow's daytime task is:

   1. check the trend data of the QPD outputs to see how much signals were there in the past.

   2. check the whitening filters to make sure if it's on or off.

   3. If it's necessary, increase the gain of the QPD to have reasonable readouts.

I am going to ask somebody to do this task.

  4299   Tue Feb 15 11:53:02 2011 kiwamuSummaryGeneraltoday's missions

I think this week is going to be an "alignment week".

The goal is to get a good alignment on X arm for both the green and the IR beam in order to reduce a2l couplings.

 

Today's missions are :

 - fixing the oplev channel names (see here)

 - fixing the oplev gain issue (see here)

 - engage the oplev servos

 - f2p adjustment

 - make a realtime lockin model for the f2p measurement and the dithering technique

 - alignment of the MC incident beam  (because we installed a new PMC this morning)

 - manual alignment of the IR beam by steering PZT1 and PZT2 (this procedure will be replaced by an automatic way soon)

 - bounce roll filters (see here)

 

  4310   Thu Feb 17 00:43:13 2011 kiwamuSummaryGeneralto do list for tomorrow

Let's share our tasks !

 

==== daytime ===

 . mode matching for MC  (Jenne/Koji)

 . mode matching for doubling crystal on PSL table (Suresh/Koji)

 . f2p adjustment (Kiwamu)

 . fix daq and CDS issues (Joe)

 . increase oplev gain (low priority)

 . make ITMY camera nicer (Steve)

 . c1ass simlink model (Valera/Joe)

 . Bounce Roll notches (Suresh)

 

==== nighttime (after 7pm) ====

 . align everything   (at first green beam, then X arm cavity and finally IR beam)

 . update the noise spectrum of the green locking

 . estimate the noise from angle to length coupling

  4317   Thu Feb 17 22:51:04 2011 josephb, valeraSummary dither alignment model

We made a model for the dither angular stabilization system c1ass.mdl. The attached file shows the diagram.

The idea is to dither a combination of 6 optics (ETMs, ITMs, PZTs) at different frequencies and demodulate three PDs (TRX, TRY, REFL11I). Then form the DOFs from demodulted signals, filter, and send each DOF to a combination of optics.

This is enough to get started with arm cavities alignment (we may need to add the BS for the Y arm). More optics and PD can be added as they become available and/or needed.

The DAC for the fast PZT  are not connected and have to be commissioned.

Attachment 1: ass-model.png
ass-model.png
  4322   Fri Feb 18 00:35:21 2011 kiwamuSummaryGeneralto do list for Feb. 18th

DAYTIME

  -  more precise F2P measurement and modify lockin simlink model (Kiwamu)

  -  run  C1ASS to check it (Valera)

  -  take care of CDS (Joe)

  -  MC mode matching (Jenne/Koji)

   -  RF stuff  (Suresh)

  -  mode matching for doubling crystal at PSL table (low priority)

  -  OPLEV (low priority)

NIGHITTIME

 -  update the noise spectra of green locking

 -  make noise budgets

  4326   Fri Feb 18 18:46:08 2011 kiwamuSummarySUSf2p done on ETMX and ITMX

The f2p measurements are done on ETMX and ITMX with the real time lockin systems.

I don't explain what is the f2p measurement in this entry, but people who are interested in it can find some details on an old elog entry here or somewhere on DCC.

So far the resultant filters looked reasonable compared with the previous SRM f2p filters.

 

- backgrounds -

  Some times ago I found that the coils on ETMX had not  been nicely balanced, and it made a POS to angle coupling when I tried green locking (see here).

In addition to that, accuracy of A2L kind of measurement including the dithering techniques depend on how well the coils are balanced.  Therefore we need to balance the coils basically at all the suspended optics.

There used to be a script for this particular purpose, called f2praio.sh. This script does measure the imbalances and then balance the coils.

However this time I used the realtime lockin system to measure the imbalances instead of using the old f2p script.

One of the reasons using the real time system is that,  some of the ezca and tds commands for the old script don't work for some reasons.

Therefore we decided to move on to the real time system like Yuta did for the A2L measurement a couple of months ago.

The f2p measurement finally gives us parameters to generate a proper set of filters for POS and also the coil gains. We apply those filters and the gains in order to eliminate the POS to angle coupling and to balance the coils.

 

- results -

The followers are the resultant filters and coil gains.

The plots below show new f2p filters according to the measurement.

f2p_ITMX.png       f2p_ETMX.png

 

ITMX (assuming pendulum POS has f0 = 1 Hz and Q = 1)

ULPOS  fz = 1.009612   Qz = 1.009612 

URPOS fz = 1.125965   Qz = 1.125965  

LLPOS  fz = 0.873725   Qz = 0.873725    

LRPOS  fz = 0.974418   Qz = 0.974418  

C1:SUS-ITMY_ULCOIL_GAIN      -1.103044

C1:SUS-ITMY_URCOIL_GAIN      0.884970

C1:SUS-ITMY_LLCOIL_GAIN      0.950650

C1:SUS-ITMY_LRCOIL_GAIN      -1.060326

 

 

ETMX (assuming pendulum POS has f0 = 1 Hz and Q = 1)

ULPOS  fz = 1.055445   Qz = 1.055445   

URPOS  fz = 1.052735   Qz = 1.052735   

LLPOS  fz = 0.944023   Qz = 0.944023   

LRPOS  fz = 0.941600   Qz = 0.941600   

C1:SUS-ETMX_ULCOIL_GAIN      -0.887550

C1:SUS-ETMX_URCOIL_GAIN      1.106585

C1:SUS-ETMX_LLCOIL_GAIN = 1.07233

C1:SUS-ETMX_LRCOIL_GAIN      -0.931013

  

The precision of the coil gains looked something like 1% because every time I ran the measurement script, the measured imbalances fluctuated at this level.

The precision of the filter gain at DC (0.01 Hz) could be worse, because the integration cycles for the measurement are fewer than that of the coil gains done at high frequency (8.5 Hz).

Of course we can make the precisions by increasing the integration cycles and the excitation amplitudes, if we want to.

  4327   Fri Feb 18 20:06:59 2011 kiwamuSummarySUScheck f2p function on ETMX

 The plot below shows how the f2p filters work.

At -2 min I turned on the f2p filters.

 f2p_ETMX.png

  4328   Fri Feb 18 20:17:07 2011 JoonhoSummaryElectronicsIsolation of Voltage regulator

Today I was working on RF distribution box.

So far I almost finished to electronically isolate voltage regulators from the box wall by inserting mica sheet, sleeve, and washers.

 

The problem I found is the resistance between wall and the voltage regulator is order of M ohms

I checked my isolation (mica sheet and sleeve and washer) but there is no problem there.

But I found that the power switch is not completely isolated from the wall.( around 800 kohm)

and that the resistance between the regulator and the wall is smaller for the regulator closer to the power switch

and greater for the regulator less closer to it.

So I think we need to put washer or sleeve to isolate the powersitch electronically from the box wall.

Suresh or I will fix this problem

[ To Suresh, I can finish the isolation when I come tomorrow. Or you can proceed to finish isolation.]

  4329   Sat Feb 19 01:58:20 2011 ranaSummaryElectronicsLow Noise BJT Pre-Amp

Frank put his low noise preamp info here.

I suggest that we build these (using Altium) but replace the cheapo transistors with the high class MAT03 matched BJT pair from Analog Devices.

This will allow us to have a pre-amp better matched to the noise of the mixers down to low frequency.

  4333   Mon Feb 21 17:29:57 2011 ranaSummaryIOOMyterious data loss: FB needs investigation

Looks like there was a mysterious loss of data overnight; since there's nothing in the elog I assume that its some kind of terrorism. I'm going to call Rolf to see if he can come in and work all night to help diagnose the issue.

Untitled.png

  4334   Mon Feb 21 23:00:06 2011 ZachSummaryelogrestarted

 again

  4343   Wed Feb 23 10:37:02 2011 josephbSummaryIOOMyterious data loss: FB needs investigation

Friday: 

In addition to the other fixes, Alex rebuilt the daqd process. I failed to elog this. When he rebuilt it, he needed change the symmerticom gps offset in the daqdrc file (located in /opt/rtcds/caltech/c1/target/fb). 

On Friday night, Kiwamu contacted me and let me know the frame builder had core dumped after a seg fault.  I had him temporarily disable the c1ass process (the only thing we changed that day), and then replaced Alex's rebuilt daqd code with the original daqd code and restarted it.  However, I did not change the symmetricom offset at this point.  Finally, I restarted the NDS process.  At that point testpoints and  trends seemed to be working.

Sunday:

The daqd process was restarted sometime on Sunday night (by Valera i believe).  Apparently this restart finally had the symmetricom gps offset kick in (perhaps because it was the first restart after the NDS was restarted?).  So data was being written to a future gps time.

Monday:

Kiwamu had problems with testpoints and trends and contacted me.  I tracked down the gps offset and fixed it, but the original daqd process only started once successfully, after that is was segfault, core dump non-stop. I tried Alex's rebuilt daqd (along with putting the gps offset to the correct value for it), and it worked.  Test points, trends, excitations were checked at the point and found working.

I still do not understand the underlying causes of all these segmentation faults with both the old and new daqd codes.  Alex has suggested some new open mx drivers be installed today.

Quote:

Looks like there was a mysterious loss of data overnight; since there's nothing in the elog I assume that its some kind of terrorism. I'm going to call Rolf to see if he can come in and work all night to help diagnose the issue.


 

  4357   Fri Feb 25 13:28:14 2011 kiwamuSummaryGeneralto do list
 Because it maybe useful to share this kind of information among us, I just put a to-do-list which has been continuously noted in my laptop.
 I guess those items can be applied for the coming two months (i.e. March and April). Anything else ?

-- Alignment

     - C1ASS

           * finalize the model and make user friendly scripts and medm screens
           * adjustment of phase rotations
           * cavity auto alignment for green

     - MC WFS 

          * put sensing matrices
          * whitening ?

    - OPLEVs

        * binary output for OPLEV whitening
        * calibration of OPLEVs  [rad/count]
        * characterization of OPLEV servos
        * bigger trans impedance gains to have reasonable readouts


-- Optimization of Suspended Optics

  - diagonalizations

         * input matrices
         * f2p with higher precision

  - damping control

        * fix ETMY damping
         * Bounce Roll notches
         * Q adjustment for damping

  - upgrading of electronics

        * bigger trans impedance gain for OSEMs
        * widen the voltage range of AA filter boards
   

   - weekly check

       * a routine script for measuring free swinging spectra


-- Input Optics

   - PMC

         * mode matching
         * epics LO HI values

   - FSS and ISS

        * recover FSS
        * make ISS working

   - EOM

       * AM minimization
       * triple resonant box

   - doubling and RFPD for green

        * mode matching to doubling crystal
        * connect RFPD_DC_MON to ADC
        * string +/-150 V and +/-15power cables from 1X1 rack to RFPDs
        * visibility check and loss investigations for the beat RFPD
        * rearrange RF amplifiers (ZLN series) for the RFPD
        * realgin Jenne's DCPD

 

-- Length Sensing and Control

   - digital system and electronics

       * characterization of RFPDs ==> SUresh/UG
       * installation of RF generation box and distribution box ==> Suresh/U
       * new LSC model and start making useful scripts (csh ? perl ? python ?)
       * binary outputs for PD whitening
       * make item lists for ordering (?)
       * draw cool diagrams for RF cable distribution and map of LSC rack.


-- Green Locking

    - X end station

           * eliminate undesired multiple spots on RFPD
           *  connect REFLPD_DCMON to ADC
           *  remote local boost
           * demodulation phase adjustment
           *  look for a high voltage amp. (bipoler)
           * installation of a mechanical shutter
           *  ETMX_TRANS CCD camera
           * analog low pass filter for temperature control

    - Y arm green locking ==> Suresh/Bryan

        * item lists
        * preparation of base mounts
        * mode matching estimation
        * mode measurement of input beam from Lightwave
        * temeperature scanning for beat location finding with IR beam
        * Installation
       * modification of PDH box

    - digital control systems

           * user friendly medm screens
           * apply proper filters for AC-DC whitened signal blending
           * add MC2 feedback path

      - noise budget

          * in-loop and out-of-loop evaluation
          * shot noise  
          * RFPD noise modeling
          * how intensity noise couples through MFD
          * electrical noise
          * frequency noise contribution from end laser and PSL
          * calibration of arm PDH signal

   - cavity scan and handing off

         * optimization of open loop transfer function for ALS
         * auto scripts   

-- misc.

   - CCD camera
        * color filters to separate IR and green (?)
   - lab laptops
        * a laptop for each end station (?)
    - dichroic TT (?)
        * large wegded and AR coated for 532nm (?)
    - epics for RS232C      
        * RS232C for doubling oven temperature

 

  4359   Fri Feb 25 14:50:16 2011 KojiSummaryGeneralto do list

- Put priority on the list

- Put names on the items

- Where is the CDS TO DO ==> Joe

-

- Remote disconnection of the greeen PDH 

- What is the situation of the PD DC for the LSC PDs?

- SUS Satelite box Resister replacement ==> Jamie

- IMC mode matching ==> Jamie/Larisa 

- Mechanical shutters everywhere

- SRM OPLEV Connection

- MC OAF

- Better LSC whitening boards

- DAFI 

Quote:
 Anything else ?

  4364   Mon Feb 28 11:22:40 2011 josephbSummaryGeneralto do list
Quote:

- Where is the CDS TO DO ==> Joe

CDS To Do:

1) Get ETMY working - figure out why signals are not getting past the AI board (D000186) to the coils.

2) Get TDS and command line AWG stuff working

3) Get c1ass and new c1lsc (with Koji) fully integrated with the rest of the system.

4) Get CDS software instructions up to date and well organized.

5) Redo cabling  and generally make it a permanent installation instead of hack job:

   a) Measure cable lengths, check connectors, wire with good routes and ensure strain relief.  Ensure proper labeling

   b) Get correct length fiber for c1sus RFM and timing.

   c) Fix up final BO adapter box and DAC boxes.

   d) Make boxes for the AA filter adapters which are currently just hanging.

   e) Get two "faceplates" for the cards in the back of the ETMY IO chassis so they can screwed down properly.

   f) Remove and properly store old, unused cables, boards, and anything else.

6) Create new documentation detailing the current 40m setup, both DCC documents and interactive wiki.

7)  Setup an Ubuntu work station using Keith's wiki instructions

 

Simulated Plant To Do:

1)Create simulated plant to interface with current end mass controls (say scx).

2) Create proper filters for pendulum and noise generation, test suspension.

3) Propagate to all other suspensions.

4) Working on simulated IFO plant to connect to LSC.  Create filters for near locked (assume initial green control perhaps) state.

5) Test LSC controls on simulated IFO.

6) Fix c code so there's seamless switching between simulated and real controls.

CDS Status:

MC damp dataviewer diaggui AWG c1lsc c1ioo c1sus c1iscex c1iscey RFM The Dolphins Sim.Plant Frame builder TDS Cabling
                             
  4371   Wed Mar 2 22:57:57 2011 sureshSummaryGeneralStuff from LLO

Here is a partial list of stuff which is being packed at LLO to be shipped to CIT.  The electronics ckt boards are yet to be added to this list.  Will do that tomorrow.

 

 

Attachment 1: eLIGO_items_from_LLO_for_Caltech.xls
  4375   Thu Mar 3 20:30:03 2011 ranaSummaryPSLPMC Sweeps @ different input power levels to measure the Finesse

Its been well noted in the past that sweeping the PMC at high power leads to a distortion of the transmitted power curve. The explanation for this was coating absorption and thermo-elastic deformation of the front face of the mirrors.

Today, I did several sweeps of the PMC. I turned off its servo and tuned its PZT so that it was nearly resonating. Then I drove the NPRO via the HV driver (gain=15) with 0-150 V (its 1.1 MHz/V) to measure the PMC transmitted light. I adjusted the NPRO pump diode current from 2A on down to see if the curves have a power dependent width.

In the picasa web slideshow:

There are 3 significant differences between this measurement and the one by John linked above: its a new PMC (Rick says its the cleanest one around), the sweep is faster - since I'm using a scope instead of the ADC I feel free to drive the thing by ~70 MHz in one cycle. In principle, we could go faster, but I don't want to get into the region where we excite the PZT resonance. Doing ~100 MHz in ~30 ms should be OK. I think it may be that going this fast avoids some of the thermal distortion problems that John and others have seen in the past. On the next iteration, we should increase the modulation index for the 35.5 MHz sidebands so as to get a higher precision calibration of the sweep's range.

By eye I find that the FWHM from image #4 is 11 ms long. That corresponds to 300 mV on the input to the HV box and 15 V on the PZT and ~16.5 MHz of frequency shift. I think we expect a number more like 4-5 MHz; measurement suspicious.

  4381   Mon Mar 7 17:58:14 2011 sureshSummaryGeneralStuff from LLO

Here is the updated list. These lists were used as packing lists and therefore are organised by Box #.
Attachment 1: eLIGO_items_from_LLO_for_Caltech_Sheet1.pdf
eLIGO_items_from_LLO_for_Caltech_Sheet1.pdf eLIGO_items_from_LLO_for_Caltech_Sheet1.pdf eLIGO_items_from_LLO_for_Caltech_Sheet1.pdf eLIGO_items_from_LLO_for_Caltech_Sheet1.pdf eLIGO_items_from_LLO_for_Caltech_Sheet1.pdf eLIGO_items_from_LLO_for_Caltech_Sheet1.pdf
  4382   Mon Mar 7 18:20:01 2011 kiwamuSummaryGreen Lockingplans
This week's goal is to investigate the source of the differential noise and to lower it.
 
Plans for tonight
 - realign GREEN_TRANS PD at the PSL table
 - update the noise budget
 - take spectrum of the differential noise
 - investigate a noise coupling to the differential noise especially from the intensity noise
 - update the noise budget again
 
Plans for this week :
 - Auto alignment scripts for green (Kiwamu)
 - connect the end REFL_DC  to an ADC (Kiwamu)
 - make an active phase rotation circuit for the end PDH (undergrads)
 - bounce-roll notches (Suresh)
 - optimization of the suspensions including the input matrices and the Q-values (Jenne)
 - optimization of MFSS (Koji/Rana/Larisa)
 - rewire the mechanical shutter on the 1X9 binary outputs (Steve)

 

  4387   Tue Mar 8 15:33:09 2011 kiwamuSummaryGreen Lockingplan on Mar.8th
Today's goal is to measure the contribution from the intensity noise to the beatnote.
 
Plans for today
  - check the ADC for the DCPD that Jenne installed yesterday
  - adjust RF power on the AOM
  - take spectrum of the differential noise and measure the coupling from the intensity noise
  - update the noise budget

Quote: from #4382
This week's goal is to investigate the source of the differential noise and to lower it.

 

  4391   Wed Mar 9 17:29:11 2011 steveSummaryVACsingle O-ring protection

We have one single O-ring on the 40m vacuum envelope. It is on the OOC west side, facing the AP table. This O-ring has to be protected from the force of this

door. There should be 3 shims  ~120 degrees apart to carry the full load, so it is not the O-ring that is getting squashed.

This morning I found only one of these shims in place.

Attachment 1: so1.jpg
so1.jpg
Attachment 2: P1070458.JPG
P1070458.JPG
  4394   Thu Mar 10 01:28:47 2011 joe, jamie, rana, chrisSummaryCDSSimSuspension !

Today was a banner day for Simulated Plants.

Joe and Jamie have been working to get it all happening and this afternoon we started stuffing filters into the plant to make it act like the:

40mETMY.png

We put in the following features so far:

  1. Anti-Imaging filters (these are hacked to be approximate since the real ones are 7570 Hz LP filters and the SimAI only can have filters up to 8192 Hz).
  2. Dewhitening filters (copied from the SimDW in the SUS-ETMY screens)
  3. Coil Driver transimpedance (1 / 200 Ohms)
  4. Magnet-coil force constant (0.016 N/A)
  5. Conversion from Coil to DOF Basis
  6. All DOFs of the mechanical model are represented as simple harmonic oscillators with Q~100 and f ~ measured free swinging peaks.
  7. Signals/Noise can be injected either as force noise on the test mass or as displacement noise at the suspension point.
  8. Conversion from DOF to Shadow Sensor basis.
  9. Optical Levers (with whitening)
  10. Shadow Sensors have 2V/mm readout gain and whitening filters before being digitized by the SimADC.

We have also changed the switching logic for the SUS and SimETMs for the shadow sensor whitening. It used to be that either the hardware OR the software whitening was on. Now we have made it like all of the other whitening/antiwhitening in LIGO and the whitening/antiwhitening come on together. Joe and Jamie are going to propagate this to the other SUS. The hardware filter is a 30,100:3 (poles:zeros) whitening filter. The digital filter used to also be 30,100:3 with a DC gain = 1. I've changed the FM1 filter in the XXSEN filter banks into a 3:30 for the ETMY so that it now comes on and just compensates the hardware filter. This change should be propagated to all other SUS and the MEDM screens updated to show the new situation.

After this change, we decided to calibrate the {UL,UR,LL,LR,SD}SEN channels into units of microns. To do this we have made an FM6 filter called 'cts2um' that accounts for the OSEM gain and the ADC conversion factors. These channels are now in units of microns without applying any calibration in the DTT or Dataviewer. The plot below shows the OSEM shadow sensor time series with all damping loops ON and a very rough version of seismic noise being injected in all 6 DOFs (note that the y-axis is microns and the x-axis is seconds).

dvsim.png

Next, Jamie is adding the angular calibrations (so that SUSPIT and SUSYAW are in rads) and Chris is making vectift quality seismic noise injectors.

We also need to add coating thermal noise, suspension thermal noise, substrate thermal noise, ADC/DAC noise and a lot of MEDM screen indicators of what state we're in. I myself can't tell from the OSEM time series if its real or Sim.

redpill_bluepill.jpg

  4412   Fri Mar 18 14:18:00 2011 kiwamuSummaryGeneralnew laser pointers

Just for a record. We got 4 new laser pointers (2 greens, 1 blue, and 1 green and red combination). Don't lose them.

They reside in a bucket on the SP table, where IR viewers and sensor cards also reside.

DSC_2883_ss.jpg

  4419   Mon Mar 21 16:49:11 2011 kiwamuSummaryGreen Lockingplan for this week

- Plan for this week

  * Intensity stabilization for the end green laser (Matt / Kiwamu)

  * Hand off the servo from Green to Red (Matt / Kiwamu)

  * Y end green locking (Suresh / Bryan) (rough schedule)

  * Reconnect the X end mechanical shutter to 1X9 (Kiwamu)

  * Connect the end DCPD signal to a DAC (done)

  * Make a LPF in a Pomona box for the temperature (Larisa)

  * Clean up and finalize the X end setup (Kiwamu)

  * Make a item lists for electronics. Order the electronics. (Aidan / Kiwamu)

  4421   Tue Mar 22 00:01:25 2011 kiwamuSummaryGreen Lockingplan for daytime tasks

Some tasks for the daytime tomorrow.

  * Beam profile measurements of the Y end laser  (Suresh / Bryan)

  * Taking care of CDS and the simulated plant (Jamie / Joe)

  * Reconnect the X end mechanical shutter to 1X9 (Kiwamu)

  * LPF for the X end temperature feedback (Larisa)

  4433   Wed Mar 23 14:19:35 2011 KojiSummaryGeneralGrand Plan

This is the grand plan we talked about in the beginning of the meeting.

  • (Kiwamu) X-end Green cleaning up / Prep for DRMI
  • (Bryan) Y-end Green
  • (Suresh) Help Bryan / RF (w. Kevin)
  • (Jenne) MC WFS / Y-arm IR alignment / MC adaptive feedforward (incl. CDS)
  • (Koji) LSC
  • (Joe) CDS cleaning up
  • (Jamie) Help Joe / Noise Budget
  • (Larisa) PMC scan / PSL photo&diagram
  • (Barbarela) ASS
  4435   Wed Mar 23 19:16:17 2011 AidanSummaryGreen LockingY-END green equipment is all available

With the exception of a 2" mirror mount, I've confirmed that we have everything for the Y-end green production and mode-matching.

We need to calculate a mode-matching solution for the Lightwave laser so that it gives the correct beam size in the doubling crystal.

Additionally, Rana has suggested that we change the pedestals from the normal 1" diameter pedestal+fork combo  to the 3/4" diameter posts and wider bases that are used on the PSL table (as shown in the attached image).

Attachment 1: three-quarter_inch_pedestal.jpg
three-quarter_inch_pedestal.jpg
  4436   Thu Mar 24 01:16:19 2011 SureshSummaryGreen LockingY-END green equipment is all available
There was a 2" mirror mount among the spares on the PSL table.  It has a window LW-3-2050 UV mounted in it.  I
have moved it to the Y-end table.  We seem to have run out of 2" mirror mounts ...
  4471   Wed Mar 30 21:43:31 2011 Aidan, KiwamuSummaryGreen LockingCalculation of the green contrast on the RF PD

Skip to final thought ...

Kiwamu and I have set about measuring the contrast of the signal on the RF PD. We can only do this when the end green laser is locked to the cavity. This is because the green transmission through the cavity, when unlocked, is too low. Unfortunately, once we lock the green beam to the cavity, we can't keep the beatnote on the RF PD stable to within a few hundred Hz of DC (remember that the cavity is swinging around by a couple of FSRs). So we also lock the PSL to cavity.

At this point we're stuck because we can't get both of these beams resonant within the cavity AND have the frequency difference between them be less 1kHz - when the lasers are locked to the cavity, their frequencies are separated by an integer number of FSRs + a fixed frequency offset, f_offset, that is set by the phase difference on reflection from the coating between the two wavelengths (532nm and 1064nm). We can never get the frequency difference between the lasers to be less than this offset frequency AND still have them both locked to the cavity.

 

So our contrast measuring method will have to use the RF signal.

 

So this is our method. We know the incident power from each beam on the RF PD (see Kiwamu's elog entry here), but to recap,

P_green_PSL = 72 uW (as measured today)

P_green_XARM = 560 uW (as measured by Kiwamu last week).

The trans-impedance of the RF PD is 240 Ohms. We'll assume a responsitivity of 0.25 A/W. So, if the XARM transmission and PSL green beams are perfectly matched then the maximum value of the RF beat note should be:

RF_amplitude_max = 2* SQRT(P_green_PSL*P_green_XARM) * responsivity * transimpedance = 240*0.25*2*(72E-6*560E-6)^(1/2) (volts)

= 24 mV = -19.5 dBm (or 27.5dBm after the +47 dB from the two  ZFL-1000LN+ amplifiers - with +15V in - that protrude from the top of the PD)

The maximum RF strength of the beat-note that we measure is around -75 dBm (at the RF output of the PD). This means the contrast is down nearly 600x from optimal. Or it means something is broken.

Final thought: at the end of this procedure we found that the RF beat note amplitude would jump to a different and much higher amplitude state. This renders a lot of the above useless until we discover the cause.

  4479   Thu Mar 31 20:37:10 2011 AidanSummaryGreen LockingRF amplitude source

 I gutted one of the $2 red laser pointers to build a laser source whose amplitude we could modulate at RF frequencies. Basically, I cut off the bulk of the housing from the pointer and soldered a BNC connection into the two terminals that the 2x 1.5V batteries were connected to. When I applied 3V across this BNC connector the diode still worked. So far so good.

Next I added a bias tee to the input. I put 3V across the DC input of the bias tee and added a -3dBm signal into the RF port of the tee. The laser beam was incident on a PDA100A (bandwidth of 1.7MHz) and, sure enough, Kiwamu and I could see a flat response in the amplitude at a given drive frequency out to around 1.7MHz.

We should check the response on a faster PD to see how fast the laser diode is, but we should be able to use this now to check the RF response of the green beat note PD. 

TO DO:

1. Add some capacitors across the DC input of the bias tee.

2. Do something about the switch on the laser diode.

3. Attach some labels to the laser that specify what is the required DC voltage and the maximum acceptable RF modulation amplitude.

Attachment 1: P1000543.jpg
P1000543.jpg
Attachment 2: P1000544.jpg
P1000544.jpg
Attachment 3: P1000545.jpg
P1000545.jpg
  4480   Thu Mar 31 20:46:11 2011 AidanSummaryGreen LockingGreen beat note PD DC response

I measured the DC response of the Green PD


Power into PD at DC (green laser pointer) = 285 uW
Voltage out of PD = 552mV/(100x SR560gain) = 5.52mV
Photocurrent = 5.52mV/(241 Ohms)*3 = 68.7uA
Responsivity = 68.7/285 = 0.24 A/W

Therefore, since the responsivity is in the correct range for a Silicon PD at 532nm, the DC output is giving us sensible response to an input signal.


But, there is a 2.12MHz, 328mV oscillation on the DC output irrespective of the incident power.
 

  4483   Fri Apr 1 23:49:24 2011 kiwamuSummaryGreen Lockingtwo states in green beat-note

According to the measurement done by Aidan and me, there are two beat-note state.

One gave us a small beat signal and the other gave us a bigger signal by approximately 20 dB.

 

 A possible reason for this phenomenon is that the end laser is operating at a special temperature that somehow drives the laser with two different modes at the same time.

So that it permits the laser sometimes locked with one of the two modes and sometimes with the other mode.

For the first step we will change the temperature such that the laser can run with a single stable mode.

Then for investigating it we will put a scanning cavity on the X end table to see if it really exhibits a two modes or not.

Quote from #4472

The attached table shows the amplitude of the green beat note when the end laser was in various states. We can increase the beat note amplitude dramatically by switching to a different states.

  4487   Tue Apr 5 17:04:36 2011 steveSummarySAFETYcranes inspected and load tested

Mike Caton of Konecranes inspected and loadtested all 3  of the 40m cranes at max reach trolley positions with 1 ton.

Attachment 1: P1070522.JPG
P1070522.JPG
Attachment 2: P1070532.JPG
P1070532.JPG
  4488   Tue Apr 5 17:31:59 2011 steveSummaryGeneralnew laser pointers

Quote:

Just for a record. We got 4 new laser pointers (2 greens, 1 blue, and 1 green and red combination). Don't lose them.

They reside in a bucket on the SP table, where IR viewers and sensor cards also reside.

DSC_2883_ss.jpg

 Low power green-red laser pointers are in. High power green, red and blue pointers are confiscated.

Attachment 1: P1070530.JPG
P1070530.JPG
  4489   Tue Apr 5 19:54:39 2011 KojiSummaryGreen LockingHamamatsu S3399 test

Since last Friday I have been testing the broadband RF photodetector in order to figure out the capability of S3399 with the similar circuit as Matt's BBPD
We also like to figure out if it has sufficient performance for the 40m green locking.

The circuit diagram is shown in the first attachment. The RF amplifier is attached at the diode while the reverse bias voltage is applied at the other side of the diode. The amplifier's input impedance is used as the transimpedance resister. Note that the bandwidth of this configuration is limited by the RC filter that consists of the junction capacitance of the diode, the series resistance of the diode, and the transimpedance resister. This cut off freq is in general lower than that cut off obtained with the usual transimpedance amplifier which has the readout resister at the feedback path of the opamp.

The transfer function of the PD is measured using Jenne's laser. At the reverse bias voltage of 30V, the -3dB bandwidth of 178MHz was obtained. This is quite high bandwidth for the most of the applications at the 40m.

Because of the low transimpedance the low-noise level of the RF amplifier is very crucial. Recently we can obtain an ultra low noise RF amplifier like Teledyne Cougar AC688 which has the NF of 0.9dB with the bandwidth between 10MHz - 600MHz. Next step will be to obtain this kind of amplifier to test the noise performance.

 

 

Attachment 1: S3399_test_110405.png
S3399_test_110405.png
Attachment 2: S3399_test_110405.pdf
S3399_test_110405.pdf
  4490   Tue Apr 5 21:20:11 2011 KojiSummaryGreen LockingX-arm cavity locked with LB1005 servo box

Last Thursday, I tested Newport Servo Controller LB1005 with the X_arm green PDH servo.

The setup and the settings I could lock the arm is depicted in the attached figure.
To lock the cavity, follow the steps below

1) Toggle the switch to the "lower" position. This disengages the servo and reset the integrator.

2) Toggle the switch to the "middle" position. The zero freq is set to the "PI corner" freq. At the low freq the gain is limited
at the value of "LF Gain Limit". This gives us a single pole at the low freq.

3) Once the lock is acquired, toggle the switch to the "upper" position. This moves the pole freq to DC, resulting in the complete integration of the signal at the low frequency.

I measured the openloop transfer function (attachment 2). The amp is quite fast and exhibits almost no phase delay upto 100kHz.
The UGF was 10kHz with the phase mergin of ~45deg. I had to tune the input offset carefully to stay at the center of the resonance.

Attachment 1: servo.png
servo.png
Attachment 2: OLTF.pdf
OLTF.pdf
  4494   Wed Apr 6 19:36:32 2011 AidanSummaryGreen Locking(In)sanity check of Green PD - some inconsistencies

I moved the Hartmut Green PD to the Jenne laser bench to try to determine if the response at RF was reasonable or somehow very much smaller than it should be. It was set up as shown in the attached diagram. The first pass at this was by comparing the ratio of the RF photocurrent of the green PD to the RF photocurrent of the New Focus 1611 InGaAs PD. That ratio (at a sufficiently low frequency) should be the same as the ratio the DC photocurrents of the two PDs.

Using the network analyzer I measured the ratio of the voltages of the two RF signals (and then scaled each of these by the respective transimpedances of the PDs: 700 Ohms for the 1611 and 240 Ohms for the Harmut PD). The resulting ratio is shown in the attached plot.

I measured the DC voltages from each PD and scaled those by the transimpedances to get the photocurrent (10 kOhm for the 1611 and 80 Ohm effective for the Harmut PD). The ratio of the DC photocurrents was 0.37. This is roughly 3x the ratio of the RF photocurrents at 500kHz (=0.14). This discrepancy is uncomfortably large.

 The full set of measurements is given in the table below:

Measurement Value
DC voltage from Hartmut PD 6.5mV (checked by turning laser on and off and measuring the difference)
DC voltage from 1611 InGaAs PD 2.20V
Transimpedance of Harmut PD at DC 80 Ohm (effective)
Transimpedance of Harmut PD at RF 240 Ohm
Transimpedance of 1611 InGaAs at DC 10 KOhm
Transimpedance of 1611 InGaAs at RF 700 Ohm
Incident Power on Hartmut PD (100% on PD area) 0.28mW (measured by Ophir power meter)
Incident Power on 1611 InGaAs (<100% on PD area) 0.64mW
Responsivity of Silicon PD at 1064nm 0.02 A/W (estimate)
Responsivity of 1611 New Focus PD at 1064nm ~0.8 A/W
   

There is one other troubling point: using the estimate of responsivity on the Harmut PD * incident power * transimpedance at DC = (0.02A/W) * (0.28mW) * (80 V/A) = 0.45 mV.

But the measured DC voltage is 6.5mV = inconsistent.

Attachment 1: PD_measurement.png
PD_measurement.png
Attachment 2: plot_PD_RF_ratios.pdf
plot_PD_RF_ratios.pdf
  4497   Thu Apr 7 11:51:13 2011 steveSummarySAFETYnew crane operator inaugurated

Quote:

Mike Caton of Konecranes inspected and loadtested all 3  of the 40m cranes at max reach trolley positions with 1 ton.

 Konecrane representative gave crane operator training in the 40m. Koji has become a qualified, trained crane operator of the 40m lab.

Attachment 1: P1070535.JPG
P1070535.JPG
  4500   Thu Apr 7 16:09:17 2011 AidanSummaryGreen Locking(In)sanity check of Green PD - some inconsistencies

I think I had underestimated the responsivity of the Silicon PD at 1064nm. The previous value was based on a rough search online for the responsivity of Silicon (I couldn't find the product number of the actual PD we are using). For instance, the PDA100A Si detector from Thorlabs has a responsivity of 0.35-0.4A/W at 1064nm. 

If we calculate the responsivity of the Hartmut PD from the measurements I made today (input power = 0.300mW, output voltage = 5.56mV, effective transimpedance = 80 Ohms), then the responsivity at 1064nm is 0.23 A/W which is not an unreasonable number given the response of the Thorlabs detector.

Quote:

Measurement Value
Responsivity of Silicon PD at 1064nm 0.02 A/W (estimate)
Responsivity of 1611 New Focus PD at 1064nm ~0.8 A/W
   

There is one other troubling point: using the estimate of responsivity on the Harmut PD * incident power * transimpedance at DC = (0.02A/W) * (0.28mW) * (80 V/A) = 0.45 mV.

But the measured DC voltage is 6.5mV = inconsistent.

 

  4501   Thu Apr 7 19:28:02 2011 KojiSummaryGreen Locking(In)sanity check of Green PD - some inconsistencies

Responsivity of SGD-444A

Quote:

For instance, the PDA100A Si detector from Thorlabs has a responsivity of 0.35-0.4A/W at 1064nm.

 

Attachment 1: SGD-444A.png
SGD-444A.png
  4502   Thu Apr 7 21:58:57 2011 AidanSummaryGreen LockingBeat note amplitude

Having convinced myself that the green Hartmut PD is giving an acceptable response at RF frequencies I decided to double-check the beatnote at IR (fiber transmission from the X-end beating with the PSL). This took a while because I had to realign the beam into the fiber at the X-end (I had a PD monitoring the output from the fiber on the PSL table and 40m of BNC cable giving me the signal from it at the X-end).

Eventually, I managed to get a beatnote on the PD. At first there was no signal at the temperature calculated using Koji and Suresh's calibration, but it turned out that the mode-overlap wasn't good enough on the PD. Now I can clearly see beats between a couple of modes, one of which is much stronger than the other. I think we should use a frequency discriminator on the output from the IR PD to servo the end laser and keep the strong beat note within <100MHz of DC.

 

  4504   Fri Apr 8 19:43:03 2011 kiwamuSummaryIOORF combiner eases impedance mismatching

An RF combiner should be included in the triple resonant box because it eases impedance mismatching and hence lowers undesired RF reflections.

Therefore we should use three cables to send the RF signals to the box and then combine them in the box.

 

(RF combiner)

 With proper terminations an RF combiner shows 50 Ohm input impedance.

But it still shows nearly 50 Ohm input impedance even if the source port is not properly terminated (i.e. non 50 Ohm termination).

This means any bad impedance mismatching on the source port can be somewhat brought close to 50 Ohm by a combiner.

  The amount of deviation from 50 Ohm in the input impedance depends on the circuit configuration of  the combiner as well as the termination impedance.

For example a resistive 3-way splitter shows 40 Ohm when the source port is shorten and the other ports are terminated with 50 Ohm.

Also it shows 62.5 Ohm when the source port is open and the other ports are terminated with 50 Ohm.

In this way an RF combiner eases  impedance mismatching on the source port.

 

(RF signal transfer at the 40m)

 According to the prototype test of the resonant box it will most likely have a non-50 Ohm input impedance at each modulation freqeucy.

If we install the resonant box apart from the combiner it will create RF reflections due to the mismatch (Case 1 in the diagram below)

The reflection creates standing waves which may excite higher harmonics and in the worst case it damages the RF sources.

 To reduce such a reflection one thing we can do is to install the combiner as a part of the resonant box (Case 2).

It will reduce the amount of the mismatching in the input impedance of the resonant circuit and results less reflections.

A rule we should remember is that a cable always needs to be impedance matched.

 

modulationBox.png

  4505   Fri Apr 8 20:43:46 2011 kiwamuSummaryIOORF combiner + resonant box : impedance and reflection coefficient

 The input impedance of the resonant box was measured when an RF combiner was attached to the box.

Indeed the combiner makes the impedance more 50 Ohm and reduces the reflection.

 

**** measurement conditions ****

* The output of box, where the EOM will be connected,  was open so that the box tries resonating with a parasitic capacitor instead of the real EOM.

* ZFSC-3-13, a 3-way combiner from mini circuit, was used.

* The S-port of the combiner was directly attached to the box with a short connector (~ 30 mm).

* Port 1 and 2 are terminated by 50 Ohm.

* The input impedance was measured on port 3 with AG4395A net work analyzer.

* Reflection coefficient 'Gamma' were calculated from the measured impedance 'Z' by using an equation Gamma = (50-Z)/(50+Z).

imp_refl.png

The resonances are found at 11, 29 and 73 MHz (55 MHz resonance was shifted to 73 MHz because of no EOM).

Note that the resonances are at frequencies where the notches appear in the reflection coefficient plot.

Don't be confused by a peak at 70 MHz in the impedance. This is an extra resonance due to a leakage inductance from the transformer in the circuit.

Quote: from #4504

An RF combiner should be included in the triple resonant box because it eases impedance mismatching and hence lowers undesired RF reflections.

  4517   Tue Apr 12 18:15:07 2011 kiwamuSummaryIOORF combiner is more like attenuator

I realized that my impedance matching theory on an RF combiner was wrong !

In fact an RF combiner behaves more like an attenuator according to a reflection measurement that I did today.

A 3-way combiner reduces power of an input signal by a factor of 4.8 dB because it can be also considered as a 3-way splitter.

So it is just a lossy component or in other words it is just an attenuator.

 

(reflection measurement)

To check my speculation that I posted on #4504 I measured reflection coefficients for both cases.

In the measurement I used a heliax cable, which goes from 1X2 rack to the PSL table with a length of about 10 m. Note that this is the cable that had been used as '33 MHz EOM'.

At the input of the heliax cable it was connected to a direction coupler to pick off reflections and the reflected signal was sampled in AG4395A.

The other end of the cable (output side of the cable) was basically connected to the resonant box.

Then I did a reflection measurement for both cases as drawn in this entry (see #4504).

  - case 1 -  the combiner was inserted at the input side of the heliax cable.

  - case 2 - the combiner was directly attached to the resonant box

On the combiner, ZFSC-3-13, the port 1 and 2 were terminated with 50 Ohm, therefore the port 3 was used as an input and the source port is the output.

Here is a resultant plot of the reflection measurements.

reflection_test.png

Note that whole data are calibrated so that it gives 0 dB when the output side of the heliax is open.

There are two things we can notice from this plot:

 (1) The reflection coefficient at the resonant frequencies (where notches appear) are the same for both cases.

 (2) Over the measured frequency range the reflections were attenuated by a factor of about 9.6 dB , which is twice as large as the insertion loss of the combiner.

These facts basically indicates that  the RF combiner behaves as a 4.8 dB attenuator.

Hence the location of the combiner doesn't change the situation in terms of RF reflections.

Quote from #4505

 The input impedance of the resonant box was measured when an RF combiner was attached to the box.

Indeed the combiner makes the impedance more 50 Ohm and reduces the reflection.

 

 

  4538   Mon Apr 18 13:05:57 2011 kiwamuSummaryLSCdemod board modification

Here is the idea how we upgrade the demodulation boards.

Basically we go ahead with two steps as depicted in the cartoon diagram below.

Once we finish the first step of upgrade, the board will be ready to install although the circuit won't be awesome in terms of noise performance.

 

demod_board.png

 

* * * (details) * * *

 First of all we will replace the home-made 90 degree splitter (see this entry) by a commercial splitter, PSCQ-2-51-W+ from Mini circuit. This is the step 1 basically.

At this point the boards will be ready to use in principle. I asked Steve to get three 90 degree splitters so that we can have at least three demodulators for the dual-recycled Michelson locking.

If they work very fine we will buy some more 90 degree splitters for full locking.

While we try to lock the dual-recycled Michelson once we will get a Cougar amplifier, remove all ERA-5s and install it such that we don't have to gain up and down in the circuit. This is the last step.

  4549   Wed Apr 20 23:20:49 2011 jamieSummaryComputersinstallation of CDS tools on pianosa

This is an overview of how I got (almost) all the CDS tools running on pianosa, the new Ubuntu 10.04 control room work station.

This is machine is experiment in minimizing the amount of custom configuration and source code compiling. I am attempting to install as many tools as possible from existing packages in

available packages

I was able to install a number of packages directly from the ubuntu archives, including fftw, grace, and ROOT:

apt-get install \
libfftw3-dev \
grace \
root-system

LSCSOFT

I installed all needed LSCSOFT packages (framecpp, libframe, metaio) from the well-maintained UWM LSCSOFT repository.

$ cat /etc/apt/sources.list.d/lscsoft.list
deb http://www.lsc-group.phys.uwm.edu/daswg/download/software/debian/ squeeze
deb-src http://www.lsc-group.phys.uwm.edu/daswg/download/software/debian/ squeeze contrib
sudo apt-get install lscsoft-archive-keyring
sudo apt-get update
sudo apt-get install ldas-tools-framecpp-dev libframe-dev libmetaio-dev lscsoft-user-en

You then need to source /opt/lscsoft/lscsoft-user-env.sh to use these packages.

EPICS

There actually appear to be a couple of projects that are trying to provide debs of EPICS. I was able to actually get epics working from one of them, but it didn't include some of the other needed packages (such as MEDM and BURT) so I fell back to using Keith's pre-build binary tarball.

Prereqs:

apt-get install \
libmotif-dev \
libxt-dev \
libxmu-dev \
libxprintutil-dev \
libxpm-dev \
libz-dev \
libxaw7-dev \
libpng-dev \
libgd2-xpm-dev \
libbz2-dev \
libssl-dev \
liblapack-dev \
gfortran

Pulled Keith's prebuild binary:

cd /ligo/apps
wget https://llocds.ligo-la.caltech.edu/daq/software/binary/apps/ubuntu/epics-3.14.10-ubuntu.tar.gz
tar zxf epics-3.14.10-ubuntu.tar.gz

GDS

I built GDS from svn, after I fixed some broken stuff [0]:

cd ~controls/src/gds
svn co https://redoubt.ligo-wa.caltech.edu/svn/gds/trunk
cd trunk
#fixed broken stuff [0]
source /opt/lscsoft/lscsoft-user-env.sh
./bootstrap
export GDSBUILD=online
export ROOTSYS=/usr
./configure --prefix=/ligo/apps/gds --enable-only-dtt --with-epics=/ligo/apps/epics-3.14.10
make
make install

dataviewer

I installed dataviewer from source:

cd ~controls/src/advLigoRTS
svn co https://redoubt.ligo-wa.caltech.edu/svn/advLigoRTS/trunk
cd trunk/src/dv
#fix stupid makefile /opt/rtapps --> /ligo/apps
make
make install

I found that the actual dataviewer wrapper script was also broken, so I made a new one:

$ cat /ligo/apps/dv/dataviewer
#!/bin/bash
export DVPATH=/ligo/apps/dv
ID=$$
DCDIR=/tmp/${ID}DC
mkdir $DCDIR
trap "rm -rf $DCDIR" EXIT
$DVPATH/dc3 -s ${NDSSERVER} -a $ID -b $DVPATH "$@"

environment

Finally, I made a environment definer file:

$ cat /ligo/apps/cds-user-env.sh
# source the lscsoft environment
. /opt/lscsoft/lscsoft-user-env.sh

# source the gds environment
. /ligo/apps/gds/etc/gds-user-env.sh

# special local epics setup
EPICS=/ligo/apps/epics
export LD_LIBRARY_PATH=${EPICS}/base/lib/linux-x86_64:$LD_LIBRARY_PATH
export LD_LIBRARY_PATH=${EPICS}/extensions/lib/linux-x86_64:$LD_LIBRARY_PATH
export LD_LIBRARY_PATH=${EPICS}/modules/seq/lib/linux-x86_64:$LD_LIBRARY_PATH
export PATH=${EPICS}/base/bin/linux-x86_64:$PATH
export PATH=${EPICS}/extensions/bin/linux-x86_64:$PATH
export PATH=${EPICS}/modules/seq/bin/linux-x86_64:$PATH

# dataviewer path
export PATH=/ligo/apps/dv:${PATH}

# specify the NDS server
export NDSSERVER=fb

[0] GDS was not compiling, because of what looked like bugs. I'm not sure why I'm the first person to catch these things. Stricter compiler?

To fix the following compile error:

TLGExport.cc:1337: error: ‘atoi’ was not declared in this scope

I made the following patch:

Index: /home/controls/src/gds/trunk/GUI/dttview/TLGExport.cc
===================================================================
--- /home/controls/src/gds/trunk/GUI/dttview/TLGExport.cc (revision 6423)
+++ /home/controls/src/gds/trunk/GUI/dttview/TLGExport.cc (working copy)
@@ -31,6 +31,7 @@
#include <iomanip>

#include <string.h>

#include <strings.h>

+#include <stdlib.h>


namespace ligogui {
using namespace std;

To fix the following compile error:

TLGPrint.cc:264: error: call of overloaded ‘abs(Int_t&)’ is ambiguous

I made the following patch:

Index: /home/controls/src/gds/trunk/GUI/dttview/TLGPrint.cc
===================================================================
--- /home/controls/src/gds/trunk/GUI/dttview/TLGPrint.cc (revision 6423)
+++ /home/controls/src/gds/trunk/GUI/dttview/TLGPrint.cc (working copy)
@@ -22,6 +22,7 @@
#include <fstream>

#include <map>
#include <cmath>

+#include <cstdlib>


namespace ligogui {
using namespace std;

  4575   Wed Apr 27 20:14:16 2011 AidanSummaryelogRestarted with script ...
  4585   Fri Apr 29 03:39:49 2011 KojiSummaryLSCCavity lengths

I tried the idea that the PRC can resonate f1 and f2 at the same time if the arm gives the reflection phase to f1 and f2 with the ratio of 1 vs 5.

The details are described on wiki. The point is this removes all of the PRC/SRC/asymmetry mumbo jumbo.

The calculated cavity lengths for f_mod of 11.065399MHz are:

  • Arm Length: 37.7974 [m]

  • PRC Length: 6.7538 [m]

  • SRC Length: 5.39915 [m]

  • Asymmetry (lx-ly): 0.0342 [m]


Here is the actual values derived from the photos.

  • Arm Length: 37.54 [m] (0.26m too short)

  • PRC Length: 6.760 [m] (6mm too long)

  • SRC Length: 5.415 [m] (16mm too long)

  • Asymmetry (lx-ly): 0.0266 [m] (8mm too long)

  4619   Tue May 3 18:25:38 2011 kiwamuSummaryLSCPRMI locking : plan

[Rana/Jamie/Kiwamu]

 Since we've got the PRMI locked we now should be able to do more qualitative measurements.

Here is a task list that we will measure/develop in the PRMI condition.

 

 

 - Optical gain measurements

 - Characterization of control loops

 - MICH and PRC calibrations

 - Noise budget

 - Development of automatic noise budget scripts

 - Arm loss measurement 

 - Shnupp asymmetry measurement

  4629   Wed May 4 15:56:09 2011 valeraSummaryGeneralPSL and MC trends

The attached plot shows 2 day trends of the PMC and MC reflected and transmitted power, the PSL POS/ANG QPD signals, and the temperature measured by the dust counter.

The power step in the middle of the plot corresponds to Koji/Jenne PMC realignment yesterday.

It looks like everything is following the day/night temperature changes.

Attachment 1: pslmcdrift.pdf
pslmcdrift.pdf