  12251   Wed Jul 6 10:58:29 2016 VarunUpdateCDSDAFI review

The DAFI block was reviewed by Rana yesterday. The following changes/improvements were suggested: (Updated on 20th July 2016 with tasks taat remain in red)

1) include all the various channels like PEM, LSC, ASC, SUS, SEI, etc. as the inputs. Currently the inputs are only the LSC.
2) include all the control signals.
3) create a very detailed diagram of the entire signal flow and plan tasks accordingly.
4) Enable cascading of various DSP processes.
5) Adjusting the gain of the AGC such that the amplitude of the output signal comes to about half the peak amplitude offered by the ADC. This will help taking advantage of the entire dynamic range of the ADC.
6) change the enable button styles from a text input based controller to a button controller.
7) Currently, disabling a particular signal terminates the signal. Instead, it should turn into a unity gain block on disabling.
8) Check if the Fibox does AC coupling or not. If not, add an AC coupling arrangement in the DAFI.
9) Check the nature of the ADC1 and ADC2 inputs to the DAFI. I checked them yesterday, and they are channels 25 and 26 of ADC0, which are empty.

  15081   Fri Dec 6 15:22:01 2019 gautamFrogsLSCDAFI system revived

[Jordan, gautam]

We did the following:

  • Route the fiber from the control room to 1Y2.
  • Plug fiber in to FiBox at either end, turned FiBoxes ON.
  • Tested the optical connection by driving a 1Vpp 440 Hz sine wave from a function generator - Yehonathan hears it loud and clear in the control room.
  • Tested that both CH1 and CH2 work - only CH1 is connected to the speakers in the control room at the moment.
  • There is some cross-coupling between the channels - not sure if this is happening in the multi-mode fiber or in the electroncis, but I estimate the isolation to be >30dB.
  • Connected CH8 and CH9 of DAC0 in the c1lsc expansion chassis to CH1 and CH2 respectively of the FiBox in 1Y2. 
  • Restarted the c1daf model on c1lsc, came up smooth.
  • Routed the POY11 error signal through the various matrices in c1daf, and we could 👂 the Y-arm cavity 🔐 😎 
  • Channels are muted for now - I'll give this a whirl while doing the PRFPMI locking.
  12150   Fri Jun 3 17:56:14 2016 VarunUpdateGeneralDAFI update

Wrote and tested a phase vocoder, with two of its applications:

1) Time scaling: This enables change of time duration without affecting the pitch.

2) Frequency warping: This changes the pitch of the sound without affecting the time duration.

1 & 2 tested offline with cavity transmission signal. 1) gives speedup of 2, and 2 gives frequency warping (pitch lowering by a factor of 2)

codes uploaded on github repo

  12154   Tue Jun 7 18:20:18 2016 VarunUpdateGeneralDAFI update

Tried to implement AGC on FE. Had some trouble bringing the code into the correct form. It looks okay now. However, this agc code as well as idenntity code (input = output) doesnt seem to build on the c1lsc FE. Have not tried too many debugging steps yet, will come and check the problem tomorrow. 

-Varun

Quote:

Wrote and tested a phase vocoder, with two of its applications:

1) Time scaling: This enables change of time duration without affecting the pitch.

2) Frequency warping: This changes the pitch of the sound without affecting the time duration.

1 & 2 tested offline with cavity transmission signal. 1) gives speedup of 2, and 2 gives frequency warping (pitch lowering by a factor of 2)

codes uploaded on github repo

 

  12159   Wed Jun 8 16:12:38 2016 VarunUpdateGeneralDAFI update

Summary: I am implementing digital audio filtering on various interferometer signals in order to listen to the processed audio which will help in characterizing and noise reduction in the interferometer. Following is an implementation of an Automatic Gain control (AGC) block on an LSC input signal.

Details of AGC: Currently, the AGC code implemented on FE takes input to fill a frame, then calculates the power in each frame and gives an appropriate gain to it, so that the new power content is to the required level. It is then written to the output, frame by frame. The frame is currently a rectangular window. The frame length and hop size can be adjusted. Current values are as follows:

frame length is 512 samples

hop length is 128 samples.

The input and output are delayed by 1 frame.

Details of testing: Attachment 1 shows a simulink diagram of the DAF system. Eric made this and I modified it later on. Testing was done using signal from the "LSC1" channel. Attachments 2 and 3 show aquired input and output of the AGC respectively. Gain of the preamp of the LSC input signal was varied over a total time span of 200 s. Each gain value was kept for a duration of about 20 seconds. The varying power levels can be seen in the input plot.

The output shows a uniform power level throughout.

 

Quote:

Tried to implement AGC on FE. Had some trouble bringing the code into the correct form. It looks okay now. However, this agc code as well as idenntity code (input = output) doesnt seem to build on the c1lsc FE. Have not tried too many debugging steps yet, will come and check the problem tomorrow. 

-Varun

Quote:

Wrote and tested a phase vocoder, with two of its applications:

1) Time scaling: This enables change of time duration without affecting the pitch.

2) Frequency warping: This changes the pitch of the sound without affecting the time duration.

1 & 2 tested offline with cavity transmission signal. 1) gives speedup of 2, and 2 gives frequency warping (pitch lowering by a factor of 2)

codes uploaded on github repo

 

 

Attachment 1: dafi.png
dafi.png
Attachment 2: agcin.pdf
agcin.pdf
Attachment 3: agcout.pdf
agcout.pdf
  12266   Thu Jul 7 12:44:52 2016 varunUpdateCDSDAFI update

Attached is a diagram, showing the entire (planned) signal flow of the DAF model. Some thoughts on the implementation after discussion with eric:

1) Since the LSC control signals and ASC signals are running on the c1lsc FE at the same rate as DAFI (16kHz), it would be wise to start from these.

   Current implementation: has a matrix at the end of the LSC PD signals, which selects one of the PD signals and outputs it to the DAFI via IPC communication.

    Proposed Changes: Add another matrix at the end of the LSC PD signals, to give to the second stereo output. Similarly, add two matrices each at the end of the LSC control signals and     the ASC signals. Each matrix must select one of the signals and output it to the DAF via IPC.

2) The PEM running on the c1sus FE system will have to be brought to DAFI in a similar fashon. However, since c1sus runs at 2kHz, there is a possibility of imaging while the signal is    transfered to the DAFI. This could be taken care of by an anti imaging filter, or inserting zeros between two samples coming at to the 16 kHz system from the 2kHz system and then low-passing it to remove the aliased parts. (similar to upsampling)

3) For the SUS control signals, input can be given from a matrix prepared for each optic seperately.

Attachment 1: DAFI.pdf
DAFI.pdf
  12324   Thu Jul 21 22:02:35 2016 varunUpdateCDSDAFI update: Frequency warping

The code for frequency warping contained a "printf()" command, which had caused the system to crash in one another instance (refer elog 12320) . Hence, I tried running the code tody by removing this line. Unfortunately, this did not work. the model still crashed. Attached is the screenshot of the FE status.

Attachment 1: 07212016.png
07212016.png
  12307   Sat Jul 16 00:30:42 2016 varunUpdateCDSDAFI update: Frequency warping | c1lsc unresponsive

Summary: I am trying to implement frequency warping/pitch shifting on the real time FE. Here is a description long overdue:

Description: The overall idea is as follows:

The DFT of a frame x_i[n]  is given by X_i[k] = \sum_{n=0}^{m-1}x_i[n]e^{-j2\pi \frac{kn}{m}}. A matrix W containing all W_{kn} = e^{-j2\pi \frac{kn}{m}} for k, n = 1, 2, ..., m can be calculated and predefined in the code. The input arrival rate is 16384 Hz, i.e. once in every 60 $\mu$s time window. Hence, the fourier coefficients can be updated cumulatively in each cycle using the current value of the input, previous value of the fourier coefficient and the components of the W matrix. This will distribute the computational load of the FFT into all the time windows. Similar operations can be carried out for the inverse STFT.

I have written and run a pseudo-real time code on my CPU. The following is the essence:
 Let the frame-length be M, and the intended scale factor of the frequency warping be 'r'. The frame overlap is 50%. At each clock cycle, the following tasks are performed:  (1 to 3 are routine tasks performed at every clock cycle, 4 is a special task performed only when a frame is filled.)
 1) Take input and apply hanning window to it.
 2) Cumulate X_i[k] for every k using the value of x_i[n] (the input) at that particular instant. Also start to cumulate X_{i+1}[k], which will be later transfered to X_i[k].
 3) Because of 4), we now have 'r+1' filled frames corresponding to output fft. Now take the ifft using two consecutive frames corresponding to only two time series points. The computations required for this task are the same as the computations required for calculation of the fourier coefficients iteratively, since the entire time series ifft is not computed.
 4) Do these special tasks after each frame gets filled:
      At this point, the ffts of the current frame and a previous frame is ready. Let us call them X1 and X2.
      Calculate phase difference between the two.
      Calculate all the interpolated |Y_i| in between these two frames depending upon the scale factor.
      Assign phase of X1 to first Y frame and assign increasing phase to all the other Y frames.
      and also do all the usual non-special tasks.

This code takes about 9-10 microseconds for a cycle with special tasks, and 5-6 microseconds for a cycle with routine tasks on my laptop (brought down from 100 microseconds peak time in the earlier offline implementation due to elemination of explicit dft and reduction in fft size), for a frame size of 32 samples. However, when fed into the c1lsc FE, it crashes, as it has done once again today evening, in the same fashon as yesterday. There could be 2 possible reasons:

1) Size of the array containing the W_{kn} matrix elements is too large for the FE memory,

2) the computations are taking up more than 60 microseconds.

Since there are already a few codes with similar array sizes, I am more inclined to think that 2) is more likely.

Another problem that I am anticipating is that for a 32 point dft and a sampling rate of 16kHz, the frequency resolution achieved is about 500 Hz, which is not sufficient if we need to represent seismic signals. The only way I can think of, for representing such signals with a small number of fft points, is to reduce the effective sampling rate, i.e. do DSP on inputs at a much lower rate than 16kHz (say 1kHz, which will give a resolution of ~30 Hz, or 2kHz giving a resolution of ~60Hz). Another advantage of this method is that it frees up more clock cycles for computation, thus the computational load can be further distributed.  The problem in this implementation is that it will increase the delays.

  12319   Thu Jul 21 12:03:35 2016 varunUpdateCDSDAFI update: Humming noise in DAFI output

 

Summary: There was always a constant humming noise in the output of speakers of both the audio channels. Tried to resolve the problem. Details are given below:

Details: The source of the noise was the typical 60 Hz (and harmonics), ~13 mV peak to peak output, in at least three channels of the DAC. (two coming from the DAF module, and one not related to the DAF.) Attachment 1 shows the noise in both the DAF channels. As compared to that, the signal coming through the AGC weak, about 6 mV RMS, about the same order as noise. In order to resolve this, the gain of the AGC was increased, so that the RMS output voltage of the Fibox (FBAO, the one at the output) was about 1.23 V RMS. It is approximately equal to +4 dBu, which is the typical expected output of the Fibox, according to the datasheet.

Attachment 1: New_Doc_13.pdf
New_Doc_13.pdf New_Doc_13.pdf
  12185   Wed Jun 15 22:12:55 2016 varunUpdateCDSDAFI update: stereo output

I wish to have stereo audio output for the DAF module. Hence, there needs to be a second output from the DAF. I added this second output to the model. Following are the details:

FiBox: It consists of two analog inputs which are digitized and multiplexed and transmitted optically. (only 1 fiber is needed due to multiplexing). Attachment 1 shows the fibox with its 2 analog inputs (one of which, is connected), and 1 fiber output. The output of the DAF goes to the FiBox. Until today, the Fibox recieved only 1 analog input. This analog signal comes from the DAC-8 (count starting from 0), which is located at "CH 1 OUT" SMA output in the "MONITORS" bin on the racks (attachment 2).

I have added another output channel to the DAF model both in software and in hardware. The DAF now also uses DAC-9 analog output which goes to the second analog input of the FiBox. The DAC-9 output is located at "CH 2 OUT" SMA output in the "MONITORS" bin on the racks (attachment 4).

After making the changes, the Fibox is shown in attacment 3.

Testing: The LSC input on passing through the DAF block is given through two different DAC outputs, to the same Fibox channel (one after the other), and the output is heard. More concrete testing will be done tomorrow. It will be as follows:

1) Currently, I need to search for a suitable cable that would connect the second channel of the output fibox to the audio mixer. After doing this, end to end testing of both channels will be done.

2) I could not access the AWG, probably because the DAQ was offline today afternoon. Using a signal from the AWG will give a more concrete testing of the stereo output.

3) After this, I will separate the two channels of the stereo completely (currectly they are seperated only at the DAF output stage)

4) I also will edit the medm gui appropriately.

 

Quote:

I have added Enable buttons for each of the DSP blocks, and labels for the matrix elements. The input matrix takes inputs from each of the 4 channels: ADC1, ADC2, LSC and EXC, and routes them to the audio processing blocks (attachment 2). The output matrix (attachment 3) takes the outputs of the various DSP blocks and routes them to the output and then to the speakers. 

 

Attachment 1: IMG_20160615_145535907.jpg
IMG_20160615_145535907.jpg
Attachment 2: IMG_20160615_145413005_HDR.jpg
IMG_20160615_145413005_HDR.jpg
Attachment 3: IMG_20160616_101229499.jpg
IMG_20160616_101229499.jpg
Attachment 4: IMG_20160616_101157096.jpg
IMG_20160616_101157096.jpg
  12211   Wed Jun 22 10:15:45 2016 varunUpdateCDSDAFI update: stereo output

I have updated the DAFI with the following changes:

1) Separated both the channels of stereo output completely, as well as in the GUI.

2) Added text monitors for the inputs and outputs.

The stereo output is now ready except for a cable going from the second channel of the output fibox to the audio mixer.

Attached is the main DAF_OVERVIEW screen and its link button from the LSC screen labelled "DAFI"

Quote:

I wish to have stereo audio output for the DAF module. Hence, there needs to be a second output from the DAF. I added this second output to the model. Following are the details:

FiBox: It consists of two analog inputs which are digitized and multiplexed and transmitted optically. (only 1 fiber is needed due to multiplexing). Attachment 1 shows the fibox with its 2 analog inputs (one of which, is connected), and 1 fiber output. The output of the DAF goes to the FiBox. Until today, the Fibox recieved only 1 analog input. This analog signal comes from the DAC-8 (count starting from 0), which is located at "CH 1 OUT" SMA output in the "MONITORS" bin on the racks (attachment 2).

I have added another output channel to the DAF model both in software and in hardware. The DAF now also uses DAC-9 analog output which goes to the second analog input of the FiBox. The DAC-9 output is located at "CH 2 OUT" SMA output in the "MONITORS" bin on the racks (attachment 4).

After making the changes, the Fibox is shown in attacment 3.

Testing: The LSC input on passing through the DAF block is given through two different DAC outputs, to the same Fibox channel (one after the other), and the output is heard. More concrete testing will be done tomorrow. It will be as follows:

1) Currently, I need to search for a suitable cable that would connect the second channel of the output fibox to the audio mixer. After doing this, end to end testing of both channels will be done.

2) I could not access the AWG, probably because the DAQ was offline today afternoon. Using a signal from the AWG will give a more concrete testing of the stereo output.

3) After this, I will separate the two channels of the stereo completely (currectly they are seperated only at the DAF output stage)

4) I also will edit the medm gui appropriately.

 

Quote:

I have added Enable buttons for each of the DSP blocks, and labels for the matrix elements. The input matrix takes inputs from each of the 4 channels: ADC1, ADC2, LSC and EXC, and routes them to the audio processing blocks (attachment 2). The output matrix (attachment 3) takes the outputs of the various DSP blocks and routes them to the output and then to the speakers. 

 

 

Attachment 1: C1DAF_OVERVIEW.png
C1DAF_OVERVIEW.png
Attachment 2: DAF_link_from_LSC.png
DAF_link_from_LSC.png
  12215   Mon Jun 27 15:12:09 2016 varunUpdateCDSDAFI update: stereo output

Using an RC to BNC connector from the inner drawer, I have added a second output cable going from the output Fibox in the control room to the audio mixer.

Quote:

I have updated the DAFI with the following changes:

1) Separated both the channels of stereo output completely, as well as in the GUI.

2) Added text monitors for the inputs and outputs.

The stereo output is now ready except for a cable going from the second channel of the output fibox to the audio mixer.

Attached is the main DAF_OVERVIEW screen and its link button from the LSC screen labelled "DAFI"

Quote:

I wish to have stereo audio output for the DAF module. Hence, there needs to be a second output from the DAF. I added this second output to the model. Following are the details:

FiBox: It consists of two analog inputs which are digitized and multiplexed and transmitted optically. (only 1 fiber is needed due to multiplexing). Attachment 1 shows the fibox with its 2 analog inputs (one of which, is connected), and 1 fiber output. The output of the DAF goes to the FiBox. Until today, the Fibox recieved only 1 analog input. This analog signal comes from the DAC-8 (count starting from 0), which is located at "CH 1 OUT" SMA output in the "MONITORS" bin on the racks (attachment 2).

I have added another output channel to the DAF model both in software and in hardware. The DAF now also uses DAC-9 analog output which goes to the second analog input of the FiBox. The DAC-9 output is located at "CH 2 OUT" SMA output in the "MONITORS" bin on the racks (attachment 4).

After making the changes, the Fibox is shown in attacment 3.

Testing: The LSC input on passing through the DAF block is given through two different DAC outputs, to the same Fibox channel (one after the other), and the output is heard. More concrete testing will be done tomorrow. It will be as follows:

1) Currently, I need to search for a suitable cable that would connect the second channel of the output fibox to the audio mixer. After doing this, end to end testing of both channels will be done.

2) I could not access the AWG, probably because the DAQ was offline today afternoon. Using a signal from the AWG will give a more concrete testing of the stereo output.

3) After this, I will separate the two channels of the stereo completely (currectly they are seperated only at the DAF output stage)

4) I also will edit the medm gui appropriately.

 

Quote:

I have added Enable buttons for each of the DSP blocks, and labels for the matrix elements. The input matrix takes inputs from each of the 4 channels: ADC1, ADC2, LSC and EXC, and routes them to the audio processing blocks (attachment 2). The output matrix (attachment 3) takes the outputs of the various DSP blocks and routes them to the output and then to the speakers. 

 

 

 

Attachment 1: IMG_20160627_151753247.jpg
IMG_20160627_151753247.jpg
  1855   Fri Aug 7 14:31:39 2009 AlbertoUpdatePSLDAQ REstarted

For some reason a few minutes ago the FB DAQ crashed and I had to restarted.

  4185   Fri Jan 21 23:17:54 2011 ranaHowToDAQDAQ Wiki Failure

The DAQ Wiki pages say to use port 8088 for restarting the Frame Builder. I tried this to no avail.

op440m:daq>telnet fb 8088
Trying 192.168.113.202...
Connected to fb.martian.
Escape character is '^]'.
^]
telnet> quit
Connection to fb.martian closed.
op440m:daq>telnet fb 8087
Trying 192.168.113.202...
Connected to fb.martian.
Escape character is '^]'.
daqd> shutdown
OK
Connection to fb.martian closed by foreign host.

Apparently, 8087 is the right port. Various elog entries from Joe and Kiwamu say 8087 or 8088. Not sure what's going on here.

After figuring this out, I activated the C1:GCV-XARM_COARSE_OUT_DAQ and C1:GCV-XARM_FINE_OUT_DAQ and set both of them to be recorded at 2048 Hz. We are loading filters and setting gains into these filter modules such that the OUT signals will be calibrated into Hz (that's why we used the OUT instead of the IN1 as there was last night).

  4194   Mon Jan 24 10:39:16 2011 josephbHowToDAQDAQ Wiki Failure

Actually both port 8087 and 8088 work to talk to the frame builder.  Don't let the lack of a daqd prompt fool you.

 

Here's putting in the commands:

rosalba:~>telnet fb 8088 Trying 192.168.113.202...

Connected to fb.martian (192.168.113.202). Escape character is '^]'.

shutdown

0000Connection closed by foreign host.

rosalba:~>date Mon Jan 24 10:30:59 PST 2011

 

Then looking at the last 3 lines of restart.log in /opt/rtcds/caltech/c1/target/fb/

daqd_start Fri Jan 21 15:20:48 PST 2011

daqd_start Fri Jan 21 23:06:38 PST 2011

daqd_start Mon Jan 24 10:30:29 PST 2011

 

So clearly its talking to the frame builder, it just doesn't have the right formatting for the prompt.  If you try typing in "help" at the prompt, you still get all the frame builder commands listed and can try using any of them.

However, I'll edit the DAQ wiki and indicate 8087 should be used because of the better formatting for the prompt.


Quote:
Apparently, 8087 is the right port. Various elog entries from Joe and Kiwamu say 8087 or 8088. Not sure what's going on here.

After figuring this out, I activated the C1:GCV-XARM_COARSE_OUT_DAQ and C1:GCV-XARM_FINE_OUT_DAQ and set both of them to be recorded at 2048 Hz. We are loading filters and setting gains into these filter modules such that the OUT signals will be calibrated into Hz (that's why we used the OUT instead of the IN1 as there was last night).

 

  4751   Thu May 19 19:41:17 2011 SureshUpdateRF SystemDAQ channel Assignments for RF PDs

[Kiwamu, Suresh]

In trying to keep the wiring of the LSC rack as neat as possible, we came up with the following channel assignments of the RF PD signals. 

PDI = PD Interface, The PD Interface D-type connectors (#1 to #12) are numbered:  Top -> Bottom and Left -> Right in ascending order. 

The Analog channels on the LSC Whitening Filter boards are numbered similarly, 1 to 32 in four sets.

1Y2_Rack_Layout.png

  3778   Mon Oct 25 20:20:59 2010 yuta, JoeUpdateCDSDAQ channel number etc
Summary:
We split the old SDSEN filters to SDSEN, SDSIDE, SDCOIL last week.
Along with this change, the TP channel number changed unfortunately.
So, we fixed them.
Also, we made FM9 do the output filter analog/digital switching.

What we did:
1. Changed the Simulink logic so that FM9 do the output filter switching, and checked the logic by probing
MAX333A for SDCOILs.

2. After making a new Simulink model and rebuilding, run the following incantation to burt restore filter
settings in /opt/rtcds/caltech/c1/target/c1SYS/c1SYSepics/ (See elog #3706)
sed -i 's/RO \(.*SW[12]R.*\)/\1/' autoBurt.req

3. DAQ channel numbers are listed in C1SYS.ini files in /cvs/cds/rtcds/caltech/c1/chans/daq/.
Channels with # signs are not activated. So, we changed, for example,

#[C1:SUS-MC1_LLSEN_IN1_DAQ]
#acquire=0
#datarate=16384
#datatype=4
#chnnum=10208

to

[C1:SUS-MC1_LLSEN_IN1_DAQ]
acquire=1
datarate=2048
datatype=4
chnnum=10208

for all channels we want to look at.

4. Restarted fb.

Plan:
- measure TFs and see if input and output filter switchings are working correctly
- make a switch that does all filter switching for all 5 OSEMS or 5 COILS
- put optical lever stuff
- fix offset sliders and offset switch
- put F2A filters in TO_COIL matrix (see elog #3769)
- make a nice graphical screen for MCs (like /cvs/cds/caltech/medm/c1/ioo/C1IOO_ModeCleaner.adl)
- write a script that activates DAQ we need
- make a plan

* SYS=SUS, RMS, MCS, etc
  3003   Fri May 28 00:40:53 2010 ranaUpdatePEMDAQ down

 Although trends are available, I am unable to get any full data from in the past (using DTT or DV). I started the FB's daqd process a few times, but no luck. 

I blame Joe's SimPlant monkeying from earlier today for lack of a better candidate. I checked and the frames are actually on the FB disk, so its something else.

  3005   Fri May 28 10:44:47 2010 josephbUpdatePEMDAQ down

Quote:

 Although trends are available, I am unable to get any full data from in the past (using DTT or DV). I started the FB's daqd process a few times, but no luck. 

I blame Joe's SimPlant monkeying from earlier today for lack of a better candidate. I checked and the frames are actually on the FB disk, so its something else.

 I tried running dataviewer and dtt this morning.  Dataviewer seemed to be working.  I was able to get trends, full data on a 2k channel (seismic channels) and full data on a 16k channel (C1:PEM-AUDIO_MIC1)  This was tried for a period 24 hours a go for a 10 minute stretch.

I also tried dtt and was able to get 2k and 16k channel data, for example C1:PEM-AUDIO_MIC1.  Was this problem fixed by someone last night or did time somehow fix it?

  3056   Tue Jun 8 18:39:36 2010 ranaUpdatePEMDAQ down

As before, I am unable to get data from the past. With DTT on Allegra I got data from now, but its unavailable from 1 hour ago. Same problem using mDV on mafalda. I blame Joe again - or the military industrial complex.

 

Quote:

Quote:

 Although trends are available, I am unable to get any full data from in the past (using DTT or DV). I started the FB's daqd process a few times, but no luck. 

I blame Joe's SimPlant monkeying from earlier today for lack of a better candidate. I checked and the frames are actually on the FB disk, so its something else.

 I tried running dataviewer and dtt this morning.  Dataviewer seemed to be working.  I was able to get trends, full data on a 2k channel (seismic channels) and full data on a 16k channel (C1:PEM-AUDIO_MIC1)  This was tried for a period 24 hours a go for a 10 minute stretch.

I also tried dtt and was able to get 2k and 16k channel data, for example C1:PEM-AUDIO_MIC1.  Was this problem fixed by someone last night or did time somehow fix it?

  6397   Fri Mar 9 20:44:24 2012 Jim LoughUpdateCDSDAQ restart with new ini file

DAQ reload/restart was performed at about 1315 PST today. The previous ini file was backed up as c1pem20120309.ini in the /chans/daq/working_backups/ directory.

I set the following to record:

The two JIMS channels at 2048:
[C1:PEM-JIMS_CH1_DQ] Persistent version of JIMS channel. When bit drops to zero indicating something bad (BLRMS threshold exceeded) happens the bit stays at zero  for >= the value of the persist EPICS variable.
[C1:PEM-JIMS_CH2_DQ] Non-persistent version of JIMS channel.

And all of the BLRMS channels at 256:
Names are of the form:
[C1:PEM-RMS_ACC1_F0p1_0p3_DQ]
[C1:PEM-RMS_ACC1_F0p3_1_DQ]

On monday I intend to look at the weekend seismic data to establish thresholds on the JIMS channels.

256 was the lowest rate possible according to the RCG manual. The JIMS channels are recorded at 2048 because I couldn't figure out how to disable the decimation filter. I will look into this further.

  6404   Tue Mar 13 13:28:31 2012 Ryan FisherUpdateCDSDAQ restart with new ini file
Extra note: This was the ini file that was edited:
/cvs/cds/rtcds/caltech/c1/chans/daq/C1PEM.ini
  16793   Thu Apr 21 10:35:23 2022 KojiUpdateCDSDAQ seemed down

Yesterday, when I worked on the damping servo, I found that any of the daqvtools (ndscope, dtt, dataviewer,...) is not available.  We may need to restart the fb and rt machines.

  4171   Thu Jan 20 00:39:22 2011 kiwamuHowToCDSDAQ setup : another trick

Here is another trick for the DAQ setup when you add a DAQ channel associated with a new front end code.

 

 Once you finish setting up the things properly according to this wiki page (this page ), you have to go to 

      /cvs/cds/rtcds/caltech/c1/target/fb

and then edit the file called master

This file contains necessary path where fb should look at, for the daqd initialization.

Add your path associated with your new front end code on this file, for example:

        /opt/rtcds/caltech/c1/chans/daq/C1LSC.ini

       /opt/rtcds/caltech/c1/target/gds/param/tpchn_c1lsc.par

After editing the file, restart the daqd on fb by the usual commands:

             telnet fb 8088

             shutdown

  16811   Mon Apr 25 17:24:06 2022 AnchalUpdateCDSDAQ still down

I investigated this issue today. At first, it seemed that only new suspension testpoints are inaccessible. I was able to use diaggui for a measurement on MC2. The DAQ network cable between 1X4 and 1Y1 was tied and is very taught now (we should relieve this as soon as possible, best solution is to lay down a longer cable over the bridge). My hypothesis is that the DAQ network might have broken while tying this cable and it probably did not come back since then.

The simplest solution would have been to restart c1su2 models. As I restarted those models though, I found that c1lsc and c1sus models failed. This is very unusual as c1su2 models are independent and share nothing with the other vertex models. So I had to restart all the FE computers eventually. But this did not solve the issue. Worse, now the DAQ isn't working for the vertex machiens as well.

Next step was to try restarting fb1 itself. We switched off all the FE computers, restarted fb1, stopped daqd_* processes, reloaded gpstime module, restarted open-mx, mx, nds and daqd_* process. But the mx.service failed to load with following error message:

● mx.service - LSB: starts MX driver for Myrinet card
   Loaded: loaded (/etc/init.d/mx)
   Active: failed (Result: exit-code) since Mon 2022-04-25 17:18:02 PDT; 1s ago
  Process: 4261 ExecStart=/etc/init.d/mx start (code=exited, status=1/FAILURE)

Apr 25 17:18:02 fb1 mx[4261]: Loading mx driver
Apr 25 17:18:02 fb1 mx[4261]: insmod: ERROR: could not insert module /opt/mx/sbin/mx_mcp.ko: File exists
Apr 25 17:18:02 fb1 mx[4261]: insmod: ERROR: could not insert module /opt/mx/sbin/mx_driver.ko: File exists
Apr 25 17:18:02 fb1 systemd[1]: mx.service: control process exited, code=exited status=1
Apr 25 17:18:02 fb1 systemd[1]: Failed to start LSB: starts MX driver for Myrinet card.
Apr 25 17:18:02 fb1 systemd[1]: Unit mx.service entered failed state.

(Ignore the timestamp above, I ran the command again to capture the error message.)

However, I was able to start all the FE models without any errors and daqd processes are also all running without showing any errors. Everything is green in CDS screen with no error messages. But the only thing still wrong is mx.service which is not running.

From my limited knowledge and experience, mx.service is a one-time script that mounts mx devices in /dev and loads the mx driver. I tried running the script /opt/mx/sbin/mx_start_stop :

controls@fb1:/opt/mx/sbin 1$ sudo ./mx_start_stop start
Loading mx driver
insmod: ERROR: could not insert module /opt/mx/sbin/mx_mcp.ko: File exists
insmod: ERROR: could not insert module /opt/mx/sbin/mx_driver.ko: File exists

This gave the same error. On searching little bit online, "insmod: ERROR; cound not insert module" error comes up when the kernel version of the driver doesnot match the Linux kernel (whatever that means!). Such deep issues should not appear out of nowhere in a previosuly perfectly runnig system. I'll check around more what changed in fb1, network cables etc.

  3631   Thu Sep 30 21:55:31 2010 ranaUpdateCDSDAQ sys update

Its pretty exciting to see that Joe got Alex to actually use the ELOG. Its a proof that even rare events occur if you are patient enough.

1) I fixed the MEDM links to point to the new sitemap.adl in /opt/rtcds. There is a link on the new sitemap which points to the old sitemap so that there is nothing destroyed yet.

2) Some of the fields in the screen are white. These are from the new c1sus processor, not issues with the slow controls. I think its just stuff that has not yet been created in the C1SUS simulink module.

Untitled.png

3) The PZT steering controls are gone. Without this we cannot get the beam down the arm. Must fix before aliging things after the MC. Since PZT used to be controlled by ASC, we'll have to wire the Piezo Jena PZT controls in from a different VME 4116. Possibly c1iool0's crate?

4) Also, the IPANG and IPPOS are somehow not working right. I guess this is because they are part of the ASC / the old ETMX system. We'll have to wire the IPANG QPD into the new ETMY ADC system if we want to get the initial alignment into the Y-arm correct.

5) I've started migrating things over from the old SITEMAP. Please just use the new SITEMAP. it has a red link to the old one, but eventually everything on the new one will work after Joe, Alex, me, and Kiwamu are done tweaking.


 

  3629   Thu Sep 30 17:11:01 2010 alex iUpdateCDSDAQ system update

The frame builder is timed from the Symmetricom GPS card now, which is getting the IRIGB timecode from the freq. distribution amplifier (from the VME GPS receiver card).

I have adjusted the GPS seconds to match the real GPS time and the DTT seems to be happy: sweeping MC2 MCL filter module produces nice plot.

Test points are working on SUS.

Excitations are working on SUS.

I am leaving the frame builder running and acquiring the data.

 

Alex

 

  3247   Mon Jul 19 21:47:36 2010 ranaSummaryDAQDAQ timing test

Since we now have a good measurement of the phase noise of the Rb clock Marconi locked to the Rb clock, I wanted to use that to check out the old DAQ system:

I used Megan's phase noise setup - Marconi #2 is putting out 11000013 Hz at 13 dBm into the ZP-3MH mixer. Marconi #1 is putting out 3 dBm at 11000000 Hz into the RF input.

The output goes through a 50 Ohm load and then a Mini-Circuits BNC LP filter (either 2 or 5 MHz). Then an SR560 set for low noise, G = 5, AC coupling, 1-pole LP @ 1 kHz.

This SR560 output goes into the channel C1:IOO-MC_DRUM1 (which is sampled at 16384 Hz with ICS-110B after the usual Sander Liu AA chassis containing the INA134s).

  3299   Tue Jul 27 16:03:36 2010 ranaSummaryDAQDAQ timing test

Quote:

Since we now have a good measurement of the phase noise of the Rb clock Marconi locked to the Rb clock, I wanted to use that to check out the old DAQ system:

I used Megan's phase noise setup - Marconi #2 is putting out 11000013 Hz at 13 dBm into the ZP-3MH mixer. Marconi #1 is putting out 3 dBm at 11000000 Hz into the RF input.

The output goes through a 50 Ohm load and then a Mini-Circuits BNC LP filter (either 2 or 5 MHz). Then an SR560 set for low noise, G = 5, AC coupling, 1-pole LP @ 1 kHz.

This SR560 output goes into the channel C1:IOO-MC_DRUM1 (which is sampled at 16384 Hz with ICS-110B after the usual Sander Liu AA chassis containing the INA134s).

 This is the 0.3 mHz BW spectrum of this test - as you can see the apparent linewidth (assuming the width is all caused by the DAQ jitter) is comparable to the BW and therefore not resolved.

Basically, the Hanning window function is not sharp enough to do this test and so I will do it offline in Matlab.

Attachment 1: Untitled.png
Untitled.png
  16853   Sat May 14 08:36:03 2022 ChrisUpdateDAQDAQ troubleshooting

I heard a rumor about a DAQ problem at the 40m.

To investigate, I tried retrieving data from some channels under C1:SUS-AS1 on the c1sus2 front end. DQ channels worked fine, testpoint channels did not. This pointed to an issue involving the communication with awgtpman. However, AWG excitations did work. So the issue seemed to be specific to the communication between daqd and awgtpman.

daqd logs were complaining of an error in the tpRequest function: error code -3/couldn't create test point handle. (Confusingly, part of the error message was buffered somewhere, and would only print after a subsequent connection to daqd was made.) This message signifies some kind of failure in setting up the RPC connection to awgtpman. A further error string is available from the system to explain the cause of the failure, but daqd does not provide it. So we have to guess...

One of the reasons an RPC connection can fail is if the server name cannot be resolved. Indeed, address lookup for c1sus2 from fb1 was broken:

$ host c1sus2
Host c1sus2 not found: 3(NXDOMAIN)

In /etc/resolv.conf on fb1 there was the following line:

search martian.113.168.192.in-addr.arpa

Changing this to search martian got address lookup on fb1 working:

$ host c1sus2
c1sus2.martian has address 192.168.113.87

But testpoints still could not be retrieved from c1sus2, even after a daqd restart.

In /etc/hosts on fb1 I found the following:

192.168.113.92  c1sus2

Changing the hardcoded address to the value returned by the nameserver (192.168.113.87) fixed the problem.

It might be even better to remove the hardcoded addresses of front ends from the hosts file, letting DNS function as the sole source of truth. But a full system restart should be performed after such a change, to ensure nothing else is broken by it. I leave that for another time.

  16854   Mon May 16 10:49:01 2022 AnchalUpdateDAQDAQ troubleshooting

[Anchal, Paco, JC]

Thanks Chris for the fix. We are able to access the testpoints now but we started facing another issue this morning, not sure how it is related to what you did.

  • The C1:LSC-TRX_OUT and C1:LSC-TRY_OUT channels are stuck to zero value.
  • These were the channels we used until last friday to align the interferometer.
  • These channels are routed through the c1rfm FE model (Reflected Memory model is the name, I think). These channels carry the IR transmission photodiode monitors at the two ends of the interferometer, where they are first logged into the local FEs as C1:SUS-ETMX_TRX and C1:SUS-ETMY_TRY .
  • These channels are then fed to C1:SCX-RFM_TRX -> C1:RFM_TRX -> C1:RFM-LSC_TRX -> C1:LSC-TRX and similar for Y side.
  • We are able to see channels in the end FE filtermodule testpoints (C1:SUS-ETMX_TRX_OUT & C1:SUS-ETMY_TRY_OUT)
  • However, we are unable to see the same signal in c1rfm filter module testpoints like C1:RFM_TRX_IN1, C1:RFM_TRY_IN1 etc
  • There is an IPC error shown in CDS FE status screen for c1rfm in c1sus. But we remember seeing this red for a long time and have been ignoring it so far as everything was working regardless.

The steps we have tried to fix this are:

  • Restart all the FE models in c1lsc, c1sus, and c1ioo (without restarting the computers themselves) , and then burt restore.
  • Restart all the FE models in c1iscex, and c1iscey (only c1iscey computer was restarted) , and then burt restore.

These above steps did not fix the issue. Since we have  the testpoints (C1:SUS-ETMX_TRX_OUT & C1:SUS-ETMY_TRY_OUT) for now to monitor the transmission levels, we are going ahead with our upgrade work without resovling this issue. Please let us know if you have any insights.

  16855   Mon May 16 12:59:27 2022 ChrisUpdateDAQDAQ troubleshooting

It looks like the RFM problem started a little after 2am on Saturday morning (attachment 1). It’s subsequent to what I did, but during a time of no apparent activity, either by me or others.

The pattern of errors on c1rfm (attachment 2) looks very much like this one previously reported by Gautam (errors on all IRFM0 ipcs). Maybe the fix described in Koji’s followup will work again (involving hard reboots).

Attachment 1: timeseries.png
timeseries.png
Attachment 2: err.png
err.png
  3057   Tue Jun 8 20:52:25 2010 josephbUpdatePEMDAQ up (for the moment)

As a test, I did a remote reboot of both Megatron and c1iscex, to make sure there was no code running that might interfere with the dataviewer.  Megatron is behind a firewall, so I don't see how it could be interfering with the frame builder.  c1iscex was only running a test module from earlier today when I was testing the multi-filter matrix part.  No daqd or similar processes were running on this machine either, but it is not behind a firewall at the moment. 

Neither of these seemed to affect the lack of past data.  I note the error message from dataviewer was "read(); errno=9".

Going to the frame builder machine, I ran dmesg.  I get some disturbing messages from May 26th and June 7th. There are 6-7 of these pairs of lines for each of these days, spread over the course of about 30 minutes.

Jun 7 14:05:09 fb ufs: [ID 213553 kern.notice] NOTICE: realloccg /: file system full

Jun 7 14:11:14 fb last message repeated 19 times

There's also one:

Jun 7 13:35:14 fb syslogd: /usr/controls/main_daqd.log: No space left on device

I went to /usr/controls/ and looked at the file.  I couldn't read it with less, it errored with Value too large for defined data type.  Turns out the file was 2.3 G.  And had not been updated since June 7th.  There were also a bunch of core dump files from May 25th, and a few more recent.  However the ones from May 25th were somewhat large, half a gig each or so.  I decided to delete the main_daqd.log file as well as the core files.

This seems to have fixed the data history for the moment (at least with one 16k channel I tested quickly). However, I'm now investigating why that log file seems to have filled up, and see if we can prevent this in the future.

Quote:

As before, I am unable to get data from the past. With DTT on Allegra I got data from now, but its unavailable from 1 hour ago. Same problem using mDV on mafalda. I blame Joe again - or the military industrial complex.

 

Quote:

Quote:

 Although trends are available, I am unable to get any full data from in the past (using DTT or DV). I started the FB's daqd process a few times, but no luck. 

I blame Joe's SimPlant monkeying from earlier today for lack of a better candidate. I checked and the frames are actually on the FB disk, so its something else.

 I tried running dataviewer and dtt this morning.  Dataviewer seemed to be working.  I was able to get trends, full data on a 2k channel (seismic channels) and full data on a 16k channel (C1:PEM-AUDIO_MIC1)  This was tried for a period 24 hours a go for a 10 minute stretch.

I also tried dtt and was able to get 2k and 16k channel data, for example C1:PEM-AUDIO_MIC1.  Was this problem fixed by someone last night or did time somehow fix it?

 

  9422   Fri Nov 22 09:54:22 2013 SteveUpdateCDSDAQ?

Jamie, I think the computers know that you are away. c1lsc keeps going down.

The short time plots are correct.

Attachment 1: comp8d.png
comp8d.png
  9423   Fri Nov 22 14:21:43 2013 JamieUpdateComputer Scripts / ProgramsDAQ?

Quote:

Jamie, I think the computers know that you are away. c1lsc keeps going down.

The short time plots are correct.

Is there some indication from the attached image that there is a problem with c1lsc?  I see some drop outs in the channels you're plotting, but those are not c1lsc channels.

The channels with the drop outs are I think derived channels, as opposed to ones that are generated on the front end.  Therefore they could have been affected by the c1auxey outages from earlier in the week.

  1737   Mon Jul 13 15:14:57 2009 AlbertoUpdateComputersDAQAWG

Today Alex came over, performed his magic rituals on the DAQAWG computer and fixed it. Now it's up and running again.

I asked him what he did, but he's not sure of what fixed it. He couldn't remember exactly but he said that he poked around, did something somewhere somehow, maybe he tinkered with tpman and eventually the computer went up again.

Now everything is fine.

  1752   Wed Jul 15 17:18:24 2009 JenneDAQComputersDAQAWG gone, now back

Yet again, the DAQAWG flipped out for an unknowable reason.  In order of restart activities listed on the Wiki, I keyed the crate and nothing really happened, then I hit the physical reset button and nothing happened, and then I did the 'telnet....vmeBusReset', and a couple minutes later, it was all good again.

  12152   Tue Jun 7 11:12:47 2016 jamieUpdateCDSDAQD UPGRADE WORK UNDERWAY

I am re-starting work on the daqd upgrade again now.  Expect the daqd to be offline for most of the day.  I will report progress.

  12155   Tue Jun 7 20:49:50 2016 jamieUpdateCDSDAQD work ongoing

Summary: new daqd code running overnight test on fb1.  Stability issues persist.

The code is from Keith's "tests/advLigoRTS-40m" branch, which is a branch of the current trunk.  It's supposed to include patches to fix the crashes when multiple frame types are written to disk at the same time.  However, the issue is not fixed:

2016-06-07_20:38:55 about to write frame @ 1149392336
2016-06-07_20:38:55 Begin Full WriteFrame()
2016-06-07_20:38:57 full frame write done in 2seconds
2016-06-07_20:39:11 about to write frame @ 1149392352
2016-06-07_20:39:11 Begin Full WriteFrame()
2016-06-07_20:39:13 full frame write done in 2seconds
2016-06-07_20:39:27 about to write frame @ 1149392368
2016-06-07_20:39:27 Begin Full WriteFrame()
2016-06-07_20:39:29 full frame write done in 2seconds
2016-06-07_20:39:43 about to write second trend frame @ 1149391800
2016-06-07_20:39:43 Begin second trend WriteFrame()
2016-06-07_20:39:43 about to write frame @ 1149392384
2016-06-07_20:39:43 Begin Full WriteFrame()
2016-06-07_20:39:44 full frame write done in 1seconds
2016-06-07_20:39:59 about to write frame @ 1149392400
2016-06-07_20:40:04 Begin Full WriteFrame()
2016-06-07_20:40:04 Second trend frame write done in 21 seconds
2016-06-07_20:40:14 [Tue Jun  7 20:40:14 2016] main profiler warning: 1 empty blocks in the buffer
2016-06-07_20:40:15 [Tue Jun  7 20:40:15 2016] main profiler warning: 0 empty blocks in the buffer
2016-06-07_20:40:16 [Tue Jun  7 20:40:16 2016] main profiler warning: 0 empty blocks in the buffer
2016-06-07_20:40:17 [Tue Jun  7 20:40:17 2016] main profiler warning: 0 empty blocks in the buffer
2016-06-07_20:40:18 [Tue Jun  7 20:40:18 2016] main profiler warning: 0 empty blocks in the buffer
2016-06-07_20:40:19 [Tue Jun  7 20:40:19 2016] main profiler warning: 0 empty blocks in the buffer
2016-06-07_20:40:20 [Tue Jun  7 20:40:20 2016] main profiler warning: 0 empty blocks in the buffer
2016-06-07_20:40:21 [Tue Jun  7 20:40:21 2016] main profiler warning: 0 empty blocks in the buffer
2016-06-07_20:40:22 [Tue Jun  7 20:40:22 2016] main profiler warning: 0 empty blocks in the buffer
2016-06-07_20:40:23 [Tue Jun  7 20:40:23 2016] main profiler warning: 0 empty blocks in the buffer

This failure comes when a full frame (1149392384+16) is written to disk at the same time as a second trend (1149391800+600).  It seems like every time this happens daqd crashes.

I have seen other stability issues as well, maybe caused by mx flakiness, or some sort of GPS time synchronization issue caused by our lack of IRIG-B cards.  I'm going to look to see if I can get the GPS issue taken care of so we take that out of the picture.

For the last couple of hours I've only seen issues with the frame writing every 20 minutes, when the full and second trend frames happen to be written at the same time.  Running overnight to gather more statistics.

  12156   Wed Jun 8 08:34:55 2016 jamieUpdateCDSDAQD work ongoing

38 restarts overnight.  Problem definitely not fixed.  I'll be reverting back to old daqd and fb this morning.  Then regroup and evaluate options.

  10481   Wed Sep 10 02:26:20 2014 JenneUpdateLSCDARM -> AS55 optickle

 Q has pointed out that we expect a sign flip in the AS55 signal for DARM as we reduce the CARM offset in the PRFPMI case.  Koji also mentioned that the SRC will help broaden the DARM linewidth.  I wanted to check and think about these things with my Optickle simulation.  Q is working on confirming my results with Mist.

The simulation situation:  

* The demod phase for AS55 is set in the MICH-only case so that the MICH signal is maximized in the Q-phase.  I do not change the demod phase at all in these simulations.

* Cavity lengths (arms, recycling cavities) are the measured lengths.

* I look at AS55 I and Q as DARM sensors (i.e. I'm doing DARM sweeps) as a function of CARM offset, for both PRFPMI and DRFPMI cases.

Spoiler alert!  Conclusions:

* In the PRFPMI case, the DARM signal shows up with approximately equal strength in the I and Q phases, so we suffer only about a factor of 2 if we do not re-optimize the demod angle for AS55.

* In the DRFPMI case, the DARM signal is a factor of 1,000 smaller in the Q-phase than the I-phase, which means that the ideal demod phase angle has moved by about 90 degrees from the MICH-only case.  We must either use the I-phase signal or change the demod phase by 90 degrees in order to acquire lock.

* In the PRFPMI case, there is a sign flip for DARM on the AS55 PD around 100pm, so we don't want to use AS55 for DARM until we are well inside 50pm, and aren't going to fluctuate out of that range.

* In the DRFPMI case, there is no such sign flip, at least out to 1nm, so we can use AS55 for DARM as soon as we see a viable signal.  This is super awesome. The caveat is that the gain changes significantly as we reduce the CARM offset, so we either need a UGF servo (eventually) or careful watching (for now).

* The AS55 linear(-ish) range is much broader in the dual recycled case, which is yet another reason why getting DARM on AS55 will be easier for DRFPMI.

Why didn't we do it already?

* To put the SRM QPD back, we'd also have to move Steve/EricG's laser.  Since I had other things to do, I left the setup for tonight, but I think I will want it for tomorrow night.

* Monday night (and tonight) we can pretty reliably get DARM onto AS55Q for the PRFPMI case, and I don't know what the cause has been for my locklosses, so I thought I'd try to figure that out first.  

Plots!

First up, the current transition we've been trying to handle, PRFPMI DARM to AS55Q.   I also plot AS55I, and we see that the signals are roughly the same magnitude (the x axis isn't the same between these plots! sorry), so we aren't screwed if we don't change the demod phase angle.  We'll be better off once we can do a re-optimization, but this is assuming we are stuck (at first) with our MICH-only demod phase angle.

AS55Q_vs_DARM_PRFPMI_0pm300pm.pngAS55I_vs_DARM_PRFPMI_0pm300pm.png

Next up, the same plots, but for the DRFPMI case.  Note here that there is a factor of about 1000 in the y-axis scales, and also that there is no switch in the sign of the zero-crossing slope for the I-phase.

AS55Q_vs_DARM_DRFPMI_0pm300pm.pngAS55I_vs_DARM_DRFPMI_0pm300pm.png

And here is the same data (DRFPMI case), but zoomed out for the Q-phase, so you can see the craziness of this phase.  Again, this is much smaller than the signals in the I-phase, so I'm not too worried.

AS55Q_vs_DARM_DRFPMI_0pm1000pm.png

Game plan:

* Steve leaves the SRM oplev back in its nominal location (we can worry about aligning the mirror, and aligning the beam on the PD, but please put it back approximately where it came from).

* Try DRMI + 2 arm locking, which I don't think we have ever actually done before.  Hopefully there won't be any tricks, and we can get to an equivalent place and successfully get DARM to AS55.  

* .... Keep going?

  1549   Tue May 5 14:02:16 2009 robUpdateLSCDARM DC response varies with DARM offset

Note the effect of quadrature rotation for small offsets.

Attachment 1: DARM_DARM_AS_DC_2.png
DARM_DARM_AS_DC_2.png
Attachment 2: DARM_DARM_AS_DC_3.png
DARM_DARM_AS_DC_3.png
Attachment 3: DARM_DARM_AS_DC_2.pdf
DARM_DARM_AS_DC_2.pdf
Attachment 4: DARM_DARM_AS_DC_3.pdf
DARM_DARM_AS_DC_3.pdf
  13799   Sun Apr 29 22:53:06 2018 gautamUpdateGeneralDARM actuation estimate

Motivation:

We'd like to know how much actuation is required on the ETMs to lock the DARM degree of freedom. The "disturbance" we are trying to cancel is the seismic driven length fluctuation of the arm cavity. In order to try and estimate what the actuation required will be, we can use data from POX/POY locks. I'd collected some data on Friday which I looked at today. Here are the results.

Method:

  • I collected the error and control signals for both arm cavities while they were locked to the PSL.
  • Knowing the POX/POY sensing response and the actuator transfer functions, we can back out the free running displacements of the two arm cavities.
    • I used numbers from the cal filters which may not be accurate (although POX sensing response which was recently measured).
    • But the spectra computed using this method seem reasonable, and the X and Y arm asds line up around 1 Hz (albeit on a log scale).
    • In this context, L_X is really a proxy for |f_X - f_{MC}| and similarly for L_Y so I think the algebra works out correctly.
    • I didn't include any of the violin mode/AA/AI filters in this calculation.
  • Having calculated the arm cavity displacements, I computed "DARM" as L_y- L_x and then plotted its asd.
  • For good measure, I also added the quadrature sum of 4 optics' displacement noise as per the 40m GWINC model - there seems to be a pretty large discrepancy, not sure why.

If this approach looks legit, I will compute the control signal that is required to stabilize this level of disturbance using the DARM control loop, and see what is the maximum permissible series resistance we can use in order to realize this stabilization. We can then compare various scenarios like different whitening schemes, with/without Barry puck etc, and look at coil driver noise levels for each of them. 

Attachment 1: darmEst.pdf
darmEst.pdf
  13805   Tue May 1 19:37:50 2018 gautamUpdateGeneralDARM actuation estimate

Here is an updated plot - the main difference is that I have added a trace that is the frequency domain wiener filter subtraction of the coherent power between the L_X and L_Y time series. I tried reproducing the calculation with the time domain wiener filter subtraction as well, using half of the time series (i.e. 5 mins) to train the wiener filter (with L_X as target and L_Y as witness), but I don't get any subtraction above 5 Hz on the half of the data that is a test data set. Probably I am not doing the pre-filtering correctly - I downsampled the signal to 1 kHz, weighted it by low passing the signal above 40 Hz and trained the Wiener filter on the resulting time series. But this frequency domain Wiener filter subtraction should be at least a lower bound on DARM - indeed, it is slightly lower everywhere than simply taking the time domain subtraction of the two data streams.

To do:

  • Re-measure calibration numbers used.
  • Redo calculation once the numbers have been verified.

Putting a slightly cleaned up version of this plot in now - I'm only including the coherence-inferred DARM estimate now instead of the straight up time domain subtraction. So this is likely to be an underestimate. At low (<10 Hz) frequencies, the time domain computation lines up fairly well, but I suspect that I am getting huge amounts of spectral leakage (see Attachment #2) in the way I compute the spectrum using scipy's filtering routine (once the Wiener filter has been computed). Note that Attachment #2, I didn't break up the data into a training/testing set as in this case, we just care about the one-off offline performance in order to get an estimate of DARM.

The python version of the wiener filter generating code only supports [b,a] output of the digital filter, an sos filter might give better results. Need to figure out the least painful way of implementing the low-noise digital filtering in python...

Attachment 1: darmEst.pdf
darmEst.pdf
Attachment 2: darmEst_time.pdf
darmEst_time.pdf
  13822   Mon May 7 16:23:06 2018 gautamUpdateGeneralDARM actuation estimate

Summary:

Using the Wiener Filter estimate of the DARM disturbance we will have to cancel, I computed how the control signal would look like for a few scenarios. Our DACs are 16-bit, +/-10V (i.e. +/-32,768cts-pk, or ~23,000cts RMS). We need to consider the shape of the de-whitening filter to conclude whether it is feasible to increase the series resistance by x10 or not.

Some details:

Note that in this first computation, I have not considered

  • Actuation range required by other loops (e.g. local damping, Oplev etc).
  • At some point, I need to add the 2P/c radiation pressure disturbance as well.
  • The control signal is calculated assuming we are actuating equally on both ETMs (but with opposite phase).
  • RMS computation is done from 30 Hz downwards, as above 30 Hz, I think the estimate from the previous elog is not true seismic displacement. 
  • De-whitening filters (or digital whitening), which will be required to suppress DAC noise at 100Hz.
  • DARM loop shape, specifically low-pass to avoid sensing noise injection. In this calculation, I just used the pendulum transfer function.

While doing this calculation, I have accounted for the fact that right now, the analog de-whitening filters in the ETM drive chain have a x3 gain which we will remove. Actually this is an assumption, I have not yet measured a transfer function, maybe I'll do one channel at EY to confirm. Also, the actuator gains themselves need to be confirmed.

As I was looking at the coil driver schematic more closely, I realized that there are actually two separate series resistances, one for the fast controls path, and another for the DC bias voltage from the slow ADCs. So I think we have been underestimating the Johnson noise of the coil drivers by sqrt(2). I've also attached screenshots of the IFOalign and MCalign screens. The two  ITMs and ETMX have pitch DC bias values that are compatible with a x10 increase of the series resistance. But even so, we will have ~3pA/rtHz per coil from the two resistances.


gautam 8pm May8: Seems like I had confirmed the x3 gain in the EX de-whitening board when Johannes and I were investigating the AI board offset.

Attachment 1: darmProj.pdf
darmProj.pdf
Attachment 2: 37.png
37.png
Attachment 3: MCalign_20180507.png
MCalign_20180507.png
  1514   Fri Apr 24 03:57:30 2009 YoichiUpdateLockingDARM demod phase
Tonight, I was able to go up to arm power = 40 by tweaking the DARM demodulation phase.
I think the DARM loop became unstable because the demodulation phase was not right and the error signal contained some junk from I-phase.
I did not do any sophisticated demodulation phase optimization. Rather I just tweaked the phase so that the dark port image becomes stable.
I will do more careful demodulation phase tuning next time.
  1515   Fri Apr 24 04:38:49 2009 YoichiUpdateLockingDARM demod phase

Quote:
Tonight, I was able to go up to arm power = 40 by tweaking the DARM demodulation phase.
I think the DARM loop became unstable because the demodulation phase was not right and the error signal contained some junk from I-phase.
I did not do any sophisticated demodulation phase optimization. Rather I just tweaked the phase so that the dark port image becomes stable.
I will do more careful demodulation phase tuning next time.


In the next try, I was actually able to go up to arm power = 70 stably.
At this power level we are ready for the RF CARM hand off.
  1516   Fri Apr 24 11:34:32 2009 robUpdateLockingDARM demod phase

Quote:

Quote:
Tonight, I was able to go up to arm power = 40 by tweaking the DARM demodulation phase.
I think the DARM loop became unstable because the demodulation phase was not right and the error signal contained some junk from I-phase.
I did not do any sophisticated demodulation phase optimization. Rather I just tweaked the phase so that the dark port image becomes stable.
I will do more careful demodulation phase tuning next time.


In the next try, I was actually able to go up to arm power = 70 stably.
At this power level we are ready for the RF CARM hand off.


There's actually code in place in the LSC to dynamically adjust the demod phase for AS1. I've never made much use of it, because it's possible to get around the problem with some gain tweaking if you start at the right phase, or because I did the DC readout handoff earlier.

Attached is a cartoon showing how the demod phase at the dark port changes as the CARM offset is decreased.
Attachment 1: darm_phase_rotate.png
darm_phase_rotate.png
  1519   Fri Apr 24 17:26:57 2009 YoichiUpdateLockingDARM demod phase

Quote:

There's actually code in place in the LSC to dynamically adjust the demod phase for AS1. I've never made much use of it, because it's possible to get around the problem with some gain tweaking if you start at the right phase, or because I did the DC readout handoff earlier.

Attached is a cartoon showing how the demod phase at the dark port changes as the CARM offset is decreased.


The cartoon is very nice.
I actually changed the demod phase continuously as the CARM offset was reduced to get up to arm power = 70.
As the CARM offset is changed, not only the DARM signal gain but also the phase margin around 100Hz changes if you use a fixed demodulation phase.
So it was necessary to change the demodulation phase to keep the DARM loop stable.
  10621   Fri Oct 17 03:05:00 2014 ericqUpdateLSCDARM locked on DC Transmission difference

 I've been able to repeatedly get off of ALS and onto (TRY-TRX)/(TRY+TRX). Nevertheless, lock is lost between arm powers of 10 and 20. 

I do the transition at the same place as the CARM->SqrtInv transition, i.e. arm powers about 1.0 Jenne started a script for the transition, and I've modified it with settings that I found to work, and integrated it into the carm_cm_up script. I've also modified carm_cm_down to zero the DARM normalization elements. 

I was thwarted repeatedly by the frequent crashing of daqd, so I was not able to take OLTFs of CARM or DARM, which would've been nice. As it was, I tuned the DARM gain by looking for gain peaking in the error signal spectrum. I also couldn't really get a good look at the lock loss events. Once the FB is behaving properly, we can learn more. 

Turning over to difference in transmission as an error signal naturally squashes the difference in arm transmissions:

TRdifflock.png


I was able to grab spectra of the error and control signals, though I did not take the time to calibrate them... We can see the high frequency sensing noise for the transmission derived signals fall as the arm power increases. The low frequency mirror motion stays about the same. 

Oct17lock.pdf


So, it seems that DARM was not the main culprit in breaking lock, but it is still gratifying to get off of ALS completely, given its out-of-loop-noise's strong dependence on PSL-alignment. 