  3179   Thu Jul 8 15:43:58 2010 ranaUpdateComputersSome channels not being recorded!!!

Quote:

This has been fixed, thanks to some help from Alex. It doesn't correspond to new computers being put in, but rather corresponds to a dcu_id change I had made in the new LSC model.

 Just as I expected, since these hunuman didn't actually check MC_L after doing this stuff, MC_L was only recording ZERO. Joe and I reset and restarted c1susmve2 and then

verified (for real this time) that the channel was visible in both the Dataviewer real time display as well as in the trend.

3-monkeys-ComicPosition.gif

The lesson here is that you NEVER trust that the problem has been fixed until you check for yourself. Also, we must always

specify a very precise test that must be used when we ask for help debugging some complicated software problem.

 

  3180   Thu Jul 8 16:24:30 2010 GopalUpdateOptic StacksCompletion of single stack layer

Single layer of stack successfully modeled in COMSOL. I'm working on trying to add Viton springs now and extend it to a full stack. Having some difficulty with finding consistent parameters to work with.

  3181   Thu Jul 8 17:29:20 2010 Katharine, SharmilaUpdateelog 


Last night, we successfully connected and powered our circuit, which allowed us to test whether our OSEMs were working.  Previously, we had been unable to accomplish this because (1) we weren't driving it sufficiently high voltage, and  (2) we didn't check that the colored leads on our circuit actually corresponded to the colored ports on the power supply (they were all switched, which we are in the process of rectifying), so our circuit was improperly connected to the supply .  Unfortunately, we didn't learn this until after nearly cooking our circuit, but luckily there appears to have been no permanent damage .

Our circuit specs suggested powering it with a voltage difference of 48V, so we needed to run our circuit at a difference of at least 36-40 V.  Since our power supply only supplied a difference of up to 30V in each terminal, we combined them in order to produce a voltage of up to double that.  We decided to power our circuit with a voltage difference of 40V (+/- 20V referenced to true ground).  The current at the terminals were 0.06 and 0.13 A. 

To test our circuit, we used a multimeter to check the supplied voltage at different test points, to confirm that an appropriate input bias was given to various circuit elements.  We identified the direction of LED bias on our OSEM, and connected it to our circuit. We were extremely gratified when we looked through the IR viewer and saw that, in fact, the LED in the OSEM was glowing happily .

P7070240.JPG P7070242.JPG

We hooked up two oscilloscopes and measured the current through the coil, and also through the LED and photodiode in the OSEM.  We observed a change in the photodiode signal when we blocked the LED light, which was expected.  The signal at the PD and the LED were both sinusoidal waves around ~3 kHz.

P7070255.JPG 

P7070257.JPG

 

We then went back to our levitation setup, and crudely tried to levitate a magnet with attached flag by using our hands and adjusting the gain (though we also could have been watching the PD current).  The first flag we tried was a soldering tip; we couldn't levitate this but achieved an interesting sort of baby-step "levitation" (levitation .15) which allowed us to balance the conical flag on its tip on top of the OSEM (stable to small disturbances).  After learning that conical flags are a poor idea, we switched our flag to a smaller-radius cylindrical magnet.  We were much closer to levitating this magnet, but were unable to conclusively levitate it .

 P7070249.JPG
Current plan:

Adjust the preset resistors to stabilize feedback

Check LED drive circuit.

Finish calculating the transfer function, and hook up the circuit to the spectrum analyzer to measure it as well.

Observe the signal from the photocurrent as disturbances block the LED light.

Play with the gain of the feedback to see how it affects levitation.

 

Attachment 1: P7070254.JPG
P7070254.JPG
  3182   Thu Jul 8 19:43:16 2010 nancyUpdateIOOWFS calculations

The WFS error signals were recorded in the order

WFS1_PIT

WFS1_YAW

WFS2_PIT

WFS2_YAW

these measurements are made in the linear region, that is the MC is nearly perfectly aligned.

This is  the average and std. dev.of 5 measurements taken of the same signals over 10 secs each. The std. dev are under 10%. And hence, I will be using 10 secs for measurements for the WFS signals after perturbations to the mirrors.

avg =

829.4408
-517.1884
297.4168
-944.7892


std_dev =

9.0506
22.9317
15.4580
8.9827

I perturbed the Pitch and Yaw of the Three mirrors (in order MC1,2,3), using ezcastep and calculated the coefficients that relate these perturbations to the WFS error signals.

The perturbation made is of -0.01 in each dof , and after measuring the WFS error for it, the system is reverted back to the previous point before making the other perturbation.

I was able to calculate the coefficients since I have assumed a linear relationship..

Following are the coefficients calculated using 10 secs measurements

coef_mat =

   1.0e+05 *

                            MC1_P   MC1_Y  MC2_P   MC2_Y    MC3_P   MC3_Y  constant
WFS1_PIT        -0.1262    0.3677   -0.4539   -0.6297   -0.1889   -0.1356   0.013664
WFS1_YAW     -0.0112   -0.7415   -0.1844    2.4509   -0.0023   -0.3531  -0.016199
WFS2_PIT         0.1251    0.4824   -0.2028   -0.6188    0.0099   -0.1490   0.006890
WFS2_YAW      0.0120   -0.7957   -0.1793    0.9962   -0.0493    0.2672 -0.013695

Also, I measured the same thing for 100s, and to my surprize, even the signs of coeficients are different.

coef_mat =


   1.0e+05 *

                           MC1_P   MC1_Y  MC2_P   MC2_Y    MC3_P   MC3_Y   constant
WFS1_PIT       -0.1981    0.3065   -0.6084   -0.9349   -0.4002   -0.3538   0.009796
WFS1_YAW     0.0607   -0.6977    0.0592    2.8753    0.3507    0.0373   -0.008194
WFS2_PIT        0.0690    0.4769   -0.2859   -0.7821   -0.1115   -0.2953  0.004150
WFS2_YAW     0.0580   -0.8153   -0.0937    1.1424    0.0650    0.4203  -0.010629

The reason I can understand is that the measurements were not made at the same time, and hence conditions might have changed.

A thing to note in all these coefficients is that they relate the error signals to the 'perturbation' around a certain point (given below). That point is assumed to lie in the linear region.

MC1_PIT      2.6129
MC1_YAW   -5.1781
MC2_PIT       3.6383
MC2_YAW    -1.2872
MC3_PIT      -1.9393
MC3_YAW    -7.518

 

  3183   Thu Jul 8 20:32:22 2010 nancyUpdateIOOMC alignment values.

I and Koji were trying to lock the mode cleaner for measuring the beam power at MC2 end. That is when we obtained the trans and refl values.

The beam characteristics at the MC2 were measured so that we could now use a dummy beam of similar power to test and characterize the QPD we are about to install at the MC2 end. This QPD wil provide two more signals in pitch and yaw, and hence complete 6 signals for 6 rotatioanl dof of the cavity. (4 are coming from WFS).

Once the QPD is characterised, it can be used to see the spot position at MC2. This is related to the mirror angles.

The width measurements were done using a beam scan. the beam scan was properly adjusted so that the maxima of the intensity of the sopt was at its center.

We also fitted gaussian curve to the beam profile, and it was a substantially good fit.

 

The whole idea is that I am trying  to look how the Wavefront sensors respond to the perturbations in the mirror angles. Once this is known, we should be able to control the mirror-movements.

the starting point would be to do just the DC measurements (which I did today). For proper analysis, AC measurements are obviously required.(will be done later).

The matrices so calculated can be inverted, and if found enough singular, the method can be used to control.

The first shot is to see teh dependency of teh error signals only on MC1 and MC3, and see if that is kind of enough to control these two mirrors.

If this works, the QPD signals could be used to control MC2 movements.

Quote:

Hmm. I expect that you will put more details of the work tomorrow.
i.e. motivation, method, result (the previous entry is only this),
and some discussiona with how to do next.

Quote:

Nancy and Koji:

This is what I and Koji measured after aligning the MC in the afternoon.

MC_Trans 4.595 (avg)

MC_Refl 0.203 (avg)

MC2_trans :

power = 1.34mW

13.5% width : x=6747.8 +- 20.7 um  , y = 6699.4+- 20.7 um

 

 

 

  3184   Thu Jul 8 21:44:43 2010 nancyUpdateIOOWFS calculations

 

I just found the singular values and the condition number of the 4*4 matrix relating the WFS error signals and the MC1 and MC2 movements.

the condition number is ~12.5. I think its small enough to continue with the scheme. (if the measurements and all are reliable).

 

  3185   Fri Jul 9 11:09:14 2010 josephbUpdateComputersFb40m and a few other machines turned off briefly just before 11am

I turned off fb40m2 and fb40m temporarily while we added an extra power strip  to the (new) 1X6 rack at the bottom in the back.  This is to allow for the addition of the 4600 computer  given to us by Rolf (which needs a good name) into the rack above the fb machine.  The fb40m2 was unfortunately plugged into the main power connectors, so we unplugged two of its cables, and put them into the new strip. While trying to undo some of the rats nest of cables in the back I also powered down and unpluged briefly the c0dcu1, the pem crate, and the myrinet bypass box.

I am in the process of bringing those machines back up and restoring the network.

Also this morning, Megatron was moved from the end station into the (new) 1X3 rack, along with its router.  This is to allow for the installation of the new end computer and IO chassis.

 

  3187   Fri Jul 9 12:07:26 2010 RazibUpdatePhase CameraWeekly update

Here are some more details about the current phasecam setup. We are using a He-Ne laser setup

phase_cam_setup_09_08_10.jpg

A crude snap shot of the setup....

mod_setup_(copy)_annotated.jpg

 

We sent light through SM2 (Steering Mirror 2)  to BS1(Beam-Splitter 1) where the laser light is split into two parts, one going to the AOM and the other to the EOM. The EOM adds 40 MHz sidebands to the incoming carrier light after SM3, and the AOM shifts the frequency of the incident light on it to 40.000 005 MHz. The purpose for doing this juggling is to intentionally create a beat signal off the reference beam from the AOM with the sidebands added at the EOM. Note that, we are driving the AOM at 7dBm and the EOM at 13 dBm with 0 (nil) modulation. The two lights are combined at the BS2 and sent off through SM5 to the camera. The CMOS of the camera contains silicon based Micro MT9V022 chip which has a quantum efficiency of 50% at 633 nm. Thus we expected a fairly good response to this He-Ne setup from the camera. 

Using a trigger circuit, we triggered the camera at 20 Hz by sending a 20Hz sinusoidal signal to it. The trigger circuit converts this to a positive square waves. Then I roughly figured out the optimum exposure time for the camera before the DC levels saturates it by writing a code for taking a series of 25 images at different exposure time. I found that 500µs seems to be a reasonable exposure time. So, using this data, I took 20 consecutive images and sent them through a short Fourier Transform segment using Matlab to see the beat signal. Note that the DC component from these processing of the images have some fringe pattern which is due to the ND 2.5 filter that we were using to reduce the light power incident on the camera. The FT method also gave us the presence of the beat signal at the corresponding bin of the FT (for example: for 5Hz beat signal, I got the DC at bin 1 of the FT and 5Hz component at bin 6 as expected). Then I changed the AOM driving frequency to 40.000 001 MHz in order to see a 1 Hz beat signal. The results for both is in my previous post. 

Quote:

I have completed the following tasks:

1. Find a simplified calibration of the exposure time for the current He-Ne setup. Basically, I triggered the camera to take 20 snapshots with a 20 Hz driving signal at different exposure time beginning from 100 us (microsecond) upto 4000 us with an increment of 200 us.

    Result: The current power allows the camera to have an exposure time of ~500 us before the DC level begans to saturate.

2. Aidan and I did some alignment and connected the AOM and corrected the driving frequency of its PZT to 40 Mhz(which apparently was disconnected which in turn gets the credit of NO beat signal for me until Tuesday 07/06/2010 5:30 PST) .

    Result: I got the beat signal of 1 Hz and 5 Hz. Image follows (the colormap shows the power in arbitrary units).

3. Finished writing my Progress Report 1 .

DC_1Hz_beat_sig.jpgDC_5Hz_beat_sig.jpg

 

  3188   Fri Jul 9 12:25:25 2010 kiwamuUpdateGreen LockingSHG on PSL table

In order to increase the green power on the PSL table, I moved the position of the Second Harmonic Generation (SHG) crystal by ~5cm.

After this modification, the green power increased from 200 uW to 640 uW. This is sufficiently good.

     As I said in the past elog entry (# 3122), the power of the green beam generated at the PSL table should be about 650 uW.

I measured the green power by the Ophir power meter and found it was ~200 uW, which made me a little bit sad.

Then I performed the beam scan measurement to confirm if the crystal  was  located on the right place. And I found the postion was off from the optimum position by ~5cm.

So I slided the postion of the SHG oven to the right place and eventually the power got increased to 640 uW.

 

some notes: 


(power measurement)

        The outgoing beam from the SHG crystal is filtered by Y1-45S to eliminate 1064nm.

According to Mott's measurement Y1 mirrors are almost transparent for green beams (T~90%), but highly reflective for 1064nm (T~0.5%).

All the green power were measured after the Y1 mirror by the Ophir configured to 532nm, though, the measured power might be offseted by a leakage of 1064nm from the Y1 mirror.

I didn't take this effect into account.

 

 

(beam scanning and positioning of crystal)

          Here is the properties of the incident beam. These numbers are derived from the beam scan measurement.

w0h             = 52.6657      +/- 0.3445 um

w0v             = 52.4798      +/- 0.1289  um

z0h             = 0.574683         +/- 0.001937 m

z0v             = 0.574325         +/- 0.0007267 m

Where the suffixes "h" and "v" stand for "horizontal" and "vertical" respectively.

The distances are calibrated such that it starts from the lens postion.

Both waist size are already sufficiently good because the optimum conversion can be achieved when the waist size is about 50um ( see this entry)

The measured data and their fitting results are shown in attachement 1.

         According to my past calculation the center of the crystal should be apart from the beam waist by 6.8mm (see this entry)

So at first I put the oven exactly on the waist postion, and then I slided it by 6.8mm.

 

 

(to be done)

        I need to find an optimum temperature for the crystal in order to maximize the green power.

Previously the optimum temperature for the crystal was 38.4 deg. But after moving the position I found the optimum temperature is shifted down to around 37deg.

Attachment 1: PSL_doubling.png
PSL_doubling.png
  3192   Mon Jul 12 10:23:51 2010 steveUpdatePEMair condition maintenance is today

The AC filters will be checked and/or replaced today. This means the AC will be off for sort periods of time. Temperature and particle count will be effected some what.

See 800 days plot

Attachment 1: pem2y.jpg
pem2y.jpg
  3195   Mon Jul 12 13:16:53 2010 kiwamuUpdateGreen LockingPZT feedback at X end

The feedback signal going to the laser PZT at the X end station was measured in the daytime and the nighttime.

It's been measured while the laser frequency was locked to the arm cavity with the green light.

arm_day_night.png

 

The red curve was measured at 3pm of 8/July Friday. And the blue curve was measured at 12am of 9/July Saturday. 

As we can see on the plot, the peak-peak values are followers

              daytime:  ~ 4Vpp

       nighttime:  ~1.8Vpp

It is obvious that the arm cavity is louder in the daytime by a factor of about 2.

Note: the feedback signal is sent to the PZT only above 1Hz because the low frequency part is stabilized mostly by the crystal temperature (see this entry)

Quote:

 What we care about is the peak-peak value of the PZT feedback signal measured on a scope for ~30 seconds.

  3197   Mon Jul 12 15:49:56 2010 nancyUpdateSUSCharacterisation of the QPD

I and koji setup the measurement of the QPD response to the pitch and yaw displacements of the beam spot.

We did this using a 100mW 1064nm laser. Its power was attenuated to ~ 1.9mW, and the spot size at the QPD position was 6000-7000 um .

The QPD was put on a translation stage, using which, the center of teh QPD wrt the beam spot could be moved in pitch and yaw.

Following are the measurements :

For yaw

:fullyaw.jpg

The slope of teh linear region is -8356 /inch

yaw_linear.jpg

 For pitch

fullpitch.jpg

The slope of the linear region in this is 9085/inch

 

pitch_linear.jpg

 

  3198   Mon Jul 12 17:05:30 2010 nancyUpdateSUSCharacterisation of the QPD

Quote:

I and koji setup the measurement of the QPD response to the pitch and yaw displacements of the beam spot.

We did this using a 100mW 1064nm laser. Its power was attenuated to ~ 1.9mW, and the spot size at the QPD position was 6000-7000 um .

The QPD was put on a translation stage, using which, the center of teh QPD wrt the beam spot could be moved in pitch and yaw.

Following are the measurements :

 

 The old plots looked horrible, and so here is a new plot

plot.png

The slopes and other stats are

Pitch

Linear model Poly1:
     f(x) = p1*x + p2
Coefficients (with 95% confidence bounds):
       p1 =        8550  (7684, 9417)
       p2 =       -2148  (-2390, -1906)

Goodness of fit:
  SSE: 9944
  R-square: 0.9923
  Adjusted R-square: 0.9907
  RMSE: 44.59

Yaw

Linear model Poly1:
     f(x) = p1*x + p2
Coefficients (with 95% confidence bounds):
       p1 =       -8310  (-8958, -7662)
       p2 =        2084  (1916, 2252)

Goodness of fit:
  SSE: 6923
  R-square: 0.9954
  Adjusted R-square: 0.9945
  RMSE: 37.21

Attachment 1: plot.png
plot.png
  3199   Mon Jul 12 18:37:10 2010 GopalUpdateOptic StacksEigenfrequency Analysis of Simple Objects

Eigenfrequency analysis has been successfully completed in COMSOL on both a tutorial camshaft, as well as a homemade metal bar.

Upon increasing in complexity to the busbar, I once again began getting into run time errors and increased lag. It seems that this is due to undefined eigenvalues when solving the linear matrices. I tried many boundary values as well as initial conditions in case this was the issue, but it was not. There seems to be some sort of an internal inconsistency. This is no longer a matter of tweaking parameters.

Next steps:

1) Try using the same techniques on the actual mirror stacks to see if we get lucky.

2) In the likely case that this doesn't happen, continue the debugging process. If necessary, a good deal of time may need to be spent learning the COMSOL lower-level jargon.

  3200   Mon Jul 12 21:26:02 2010 Katharine, SharmilaUpdateelogmaglev coils

The connection between our coil wires and BNC terminals was pretty awful (soldered wires broke off ) so we removed the old heat shrink and re-soldered the wires.  We then chose more appropriately sized heat shrinks (small one around each of the two soldered wires, a medium-sized shrink around the wires together, a large one covering the BNC terminal and the wire) and used the solder iron and heat gun to shrink them.

P7120276.JPG

 

 

  3201   Mon Jul 12 22:01:13 2010 KojiUpdateSUSOne TT suspended. Still need fine alignment

Jenne and Koji

We tweaked the alignment of the TT mirror.

First we put a G&H mirror, but the mirror was misaligned and touching the ECD as the magnet was too heavy. We tried to move the wires towards the magnet by 1mm.
It was not enough but once we moved the clamps towards the magnet, we got the range to adjust the pitching back and forth.
We tried to align it by the feaher touch to the clamp, we could not get close to the precision of 10mrad as the final tightening of the clamp screws did change the alignment.

We will try to adjust the fine alignment tomorrow again.

The damping in pitch, yaw and longitudinal looks quite good. We will also try to characterize the damping of the suspension using a simple oplev setup.

Attachment 1: IMG_2634.jpg
IMG_2634.jpg
  3202   Tue Jul 13 10:02:30 2010 steveUpdateMOPAlaser power is dropping slowly

I have just removed an other 400 cc of water from the chiller.  I have been doing this since the HTEMP started fluctuating.

The Neslab bath temp is 20.7C, control room temp 71F

 

Attachment 1: power100d.jpg
power100d.jpg
  3203   Tue Jul 13 11:00:29 2010 kiwamuUpdateGreen LockingSHG on PSL table : optimum temeprature

The optimum temperature for the doubling crystal on the PSL table was found to be 36.8 deg.

I scanned the temperature of the crystal from 44 deg to 29 deg, in order to find the optimum temperature where the frequency doubled power is maximized. 

 

(method) 

The method I performed is essentially the same as that Koji did before (see this entry).

(1) First of all, I enabled the PID control on the temperature controller TC200 and set the temperature to 44 deg.

(2) After it got 44 deg, I disabled the PID control.

(3) Due to the passive cooling of the oven, the temperature gradually and slowly decreased. So it automatically scans the temperature down to the room temperature.

(4) I recorded the power readout of the power meter: New Port 840 together with the temperature readout of TC200. The power meter was surely configured for 532 nm.

 

(result)

The measured data are shown in the attachment. 

The peak was found at T=36.8 deg where the power readout of  532 nm was 605 uW.

Compared with Koji's past data (see this entry), there are no big side lobes in this data. I am not sure about the reason, but the side lobes are not critical for our operation of the green locking.

 

 (to be done)

 Adjustment of the PID parameters

Attachment 1: power_temp.png
power_temp.png
  3204   Tue Jul 13 11:20:07 2010 DmassUpdateGreen LockingSHG on PSL table : optimum temeprature

 

 It seems like you might inherit an offset by using step (3) b/c of the temperature gradient between the crystal and the sensing point. Depending on how large this gradient is you could increase the linear coupling from temperature to intensity noise from zero to a significant number. Phase noise should not be effected.

SInce these things (ovens) are so low time constant, shouldn't we

  1. Lock to a temperature
  2. Let the oven equilibrate for however long - a few tau maybe - my oven has a time constant of 60 sec, don't know if this is fast or slow compared to that
  3. Measure P_532/P_1064
  4. Change the setpoint
  5. Go back to step 1
  3205   Tue Jul 13 11:51:19 2010 JenneUpdatePEMGur1 has no signal?!?!?!?!

[Sanjit, Jenne]

Sanjit discovered that the Gur1 channels are all digital 0.  We determined that this began on July8, 04:00 UTC (~9pm on the 7th?).

It's digital zero, so we suspect a software thing.  Just to check, we put a sine wave in, and didn't see anything.  Gur2 seems totally fine, and the sine wave input showed up nicely on dataviewer.  What's going on? Sabotage to prevent this paper from getting done?  Dmass trying to get his paper done before me???

Investigations are ongoing.... Joe claims it's not his fault, since his shenanigans near the PEM rack were on days before, and days after this, but not on the 7th.  

Attachment 1: Gur1_noSignal_13Jul2010.png
Gur1_noSignal_13Jul2010.png
  3207   Tue Jul 13 14:59:04 2010 GopalUpdateOptic StacksEigenfrequency Analysis of Single Stack Complete

Via reconfiguration of Viton parameters (previously posted), I managed to debug the COMSOL run time errors and null pointer exceptions. Listed are the resultant eigenfrequencies obtained through structural analysis testing. For all tests, the bottom of the Viton springs are constrained from motion, and all other parts are free to oscillate. Notice that color variations signify displacement from the equilibrium position. Also note that different initial conditions produce different eigenmodes:

No initial displacement:

Eigenmode_Stack_4.png


0.01 m x-displacement:

Eigenmode_Stack_4_xdisp.png


0.01 m y-displacement:

Eigenmode_Stack_4_ydisp.png


 0.01 m z-displacement:

 Eigenmode_Stack_4_zdisp.png


Clearly, the plate has its first harmonic between 210-215 Hz, which is much greater than seismic noises (which never exceed the 10-Hz range). This suggests a highly attenuating transfer function. Since the remaining three plates have been designed to resonate similarly, it is likely that the entire stack system will also function very well.

Next steps:

1) Extend the eigenfrequency analysis to obtain a transfer function for the single-plate system

2) Expand the CAD model to include all four stack layers, and perhaps a base

 

  3208   Tue Jul 13 17:36:42 2010 nancyUpdateIOOWavefront Sensing Matrix Control

For yesterday - July 12th.

Yesterday, I tried understanding the MEDM and the Dataviewer screens for the WFS.

I then also decided to play around with the sensing matrix put into the WFS control system and see what happens.

I changed the sensing matrix to completely random values, and for some of the very bad values, it even lost lock :P (i wanted that to happen)

Then I put in some values near to what it already had, and saw things again.

I also put in the matrix values that I had obtained from my DC calculations, which after Rana's explanation, I understand was silly.

Later I put back the original values, but the MC lock didnot come back to what it was earlier. Probably my changing the values took it out of the linear region. THE MATRIX NOW HAS ITS OLD VALUES.

I was observing the POwer Spectrum of teh WFS signals after changing the matrix values, but it turned out to  be a flop, because  I had not removed the mean while measuring them.  I will do that again today, if we obtain the lock again (we suddenly lost MC lock badly some 20 minutes ago).

  3209   Tue Jul 13 19:26:47 2010 ranaUpdateIOOMC still broken

After whatever Joe/Alberto did this afternoon, the MC was not locking. Koji and I removed several of the cables in the side of the rack where they

were apparently working (I say apparently because there's no elog).

MC is now locking but the autolocker did not work at first - op340m was unable to access any channels from c1iool0. After several minutes, it mysteriously

started working - the startup.cmd yields errors seen on the terminal. I attach the screen dump/.

Attachment 1: a.txt.gz
  3211   Tue Jul 13 23:20:37 2010 JenneUpdatePEMGur1 has no signal?!?!?!?!

Quote:

[Sanjit, Jenne]

Sanjit discovered that the Gur1 channels are all digital 0.  We determined that this began on July8, 04:00 UTC (~9pm on the 7th?).

It's digital zero, so we suspect a software thing.  Just to check, we put a sine wave in, and didn't see anything.  Gur2 seems totally fine, and the sine wave input showed up nicely on dataviewer.  What's going on? Sabotage to prevent this paper from getting done?  Dmass trying to get his paper done before me???

Investigations are ongoing.... Joe claims it's not his fault, since his shenanigans near the PEM rack were on days before, and days after this, but not on the 7th.  

 I just rebooted c0dcu1, which didn't help anything.  Joe said he'd try to give me a hand tomorrow.

  3213   Wed Jul 14 10:00:14 2010 josephbUpdatePhase CameraWork near 1Y2 yesterday

Razib and I were attempting to get the output of a photodiode (PD55A in this case) recorded, so that we could independently measure the slow (~1-10 Hz) fluctuations of the light incident on the camera.  This would then allow us to subtract those fluctuations out, letting us get at the camera noise in the case with signal present (as opposed to just a dark noise measurement when we look at the noise with no signal present).

Originally I was thinking of using one empty patch panel BNCs used for PEM channels down by the 1Y7 rack and go through a 110B, although Alberto pointed out he had recently removed some monitoring equipment, which watched the amplitude modulation at various frequencies of the RF distribution (i.e. 33 MHz, etc).  This equipment output a DC voltage proportional to the amplitude of the RF signals.  The associated channel names were C1:IOO-RFAMPD_33MHZ, C1:IOO-RFAMPD_33MHZ_CAL, C1:IOO-RFAMPD_133MHZ, etc.  These are slow channels, so I presume they enter in via the slow computers, probably via pentek (I didn't check that, although in hindsight I probably should have taken the time to find exactly where they enter the system).  The connections them selves were a set of BNCs on the south side, half way up the 1Y2 rack.

We simply chose one, the 33 MHz channel in this case, and connected.  At around this time, the MC did become unlocked, although it looked like it was due to the MC2 watchdog tripping.  The initial theory was we had bumped the Mode Cleaner while looking around for some BNC cables, although from what Rana had to do last night, it probably was the connection.  We were able to restore the watchdog and confirm that the optic started to settle down again.  Unfortunately, I had to leave about 5 minutes later, and didn't do as thorough an investigation as was warranted.

  3214   Wed Jul 14 11:32:36 2010 Sharmila, KatharineUpdateWIKI-40M Update 

Weekly update:

We correctly connected our circuit to power source to verify that it was functional and that our LED in the shadow sensor turned on.  It did, but we also noticed a funky signal from the LED driver.  We continued to attempt 1x1 levitation, but determined that the temporary flag we were using out of convenience (a long, thin cylindrical magnet) was weakly attracted to residually magnetized OSEM components.  We then switched to an aluminum screw as our flag.

We resoldered and applied heat shrink to the wires connecting our coil to the BNC terminal, since they were falling apart.

We sat down with Rana and removed circuit components in the LED drive part by part to determine what was tripping up the circuit.  We determined a rogue capacitor to be at fault and removed it from the circuit.

We used a spectrum analyzer to measure the frequency response of our circuit (see details in last elog).  We are currently making a Simulink block diagram so we can check the stability of our setup, but are temporarily set back because our plotted calculation of the transfer function clearly doesn't match the measured one.

  3215   Wed Jul 14 11:51:48 2010 RazibUpdatePhase CameraWork near 1Y2 yesterday

Quote:

Razib and I were attempting to get the output of a photodiode (PD55A in this case) recorded, so that we could independently measure the slow (~1-10 Hz) fluctuations of the light incident on the camera.  This would then allow us to subtract those fluctuations out, letting us get at the camera noise in the case with signal present (as opposed to just a dark noise measurement when we look at the noise with no signal present).

Originally I was thinking of using one empty patch panel BNCs used for PEM channels down by the 1Y7 rack and go through a 110B, although Alberto pointed out he had recently removed some monitoring equipment, which watched the amplitude modulation at various frequencies of the RF distribution (i.e. 33 MHz, etc).  This equipment output a DC voltage proportional to the amplitude of the RF signals.  The associated channel names were C1:IOO-RFAMPD_33MHZ, C1:IOO-RFAMPD_33MHZ_CAL, C1:IOO-RFAMPD_133MHZ, etc.  These are slow channels, so I presume they enter in via the slow computers, probably via pentek (I didn't check that, although in hindsight I probably should have taken the time to find exactly where they enter the system).  The connections them selves were a set of BNCs on the south side, half way up the 1Y2 rack.

We simply chose one, the 33 MHz channel in this case, and connected.  At around this time, the MC did become unlocked, although it looked like it was due to the MC2 watchdog tripping.  The initial theory was we had bumped the Mode Cleaner while looking around for some BNC cables, although from what Rana had to do last night, it probably was the connection.  We were able to restore the watchdog and confirm that the optic started to settle down again.  Unfortunately, I had to leave about 5 minutes later, and didn't do as thorough an investigation as was warranted.

 Before I left, I disconnected the PD55, so the 33 MHz channel wasn't physically connected to anything!!! Only one end of the wire was connected to the rack while the other was free...

So it wasn't the PD connection that is responsible for MC tripping at the later time...

  3216   Wed Jul 14 11:54:33 2010 josephbUpdateDAQDebugging Guralp and reboots

This is regards to zero signal being reported by the channels C1:PEM-SEIS_GUR1_X, C1:PEM-SEIS_GUR1_Y, and C1:PEM-SEIS_GUR1_Z.

I briefly swapped Guralp 1 EW and Guralp 2 EW to confirm to myself that it was not on the gurlap end (although the fact that its digital zero is highly indicative a digital realm problem).  I then unplugged the 17-32, and then 1-16 channel connections to the 110B.  I saw floating noise on the GUR2 channels, but still digital zero on the GUR1 channels, which means its not the BNC break out box.

There was a spare 110B, unconnected in the crate, so to do a quick test of the 110B, I turned off the crate and swapped the 110Bs, after copying the switch configuration of the first 110B to the second one.  The original 110B was labeled ADC 1, while the second 110B was labeled ADC 0.  The switches were identical except for the ones closest to the Dsub connectors on the front.  All those switches in that set were to the right, when looking down at the switches and the Dsub connectors pointing towards yourself.

Unfortunately, the c0duc1 never seemed to come up with the new 110B (ADC 0).  So we put the original 110B back.  And turned the crate back on. 

The fb then didn't seem to come back quite right.  We tried rebooting fb40m it, but its still red with status 1.  c0daqctrl is green, but c0dcu1 is red, although I'm not positive if thats due to fb40m being in a strange state.  Jenne tried a telnet in to port 8087 and shutdown, but that didn't seem to help.  At this point, we're going to contact Alex when he gets in around 12:30.

 

  3218   Wed Jul 14 12:31:11 2010 nancyUpdateGeneralWeekly Update

Summary of this week's work Wednesday - Aligned the mode cleaner with Koji, and then measured the beam characteristics at MC2 end. Koji then taught me how to read the WFS signals Thursday - wrote a script to measure the signals and calculated the coefficients relating mirror movement and DC signals of WFS. To know the possibility of the control, found SVD of the coeff matrix, and condition number. Friday - Set up the measurement of QPD linear response using a laser outside the cavity. Took data. Monday - did the calculations and plotting for the above experiment. Then played around with the MEDM screens , and also tried to see what happens to the Power Spectrum of WFS signals by changing the coefficients in teh matrix. (failed) Tuesday - played around with WFS, tried seeing what it does when switched on at different points, and also what it does when I disturb the system while WFS has kept it locked.

Today - had switched off the WFS sensors yesterday night after locking the MC as wanted to know that how does MC behave when no WFS gain is applied. I checked in the morning, the MC was locked all night. I am now proceding with my calculations for the sensing matrix
  3219   Wed Jul 14 13:03:04 2010 Gopal UpdateWIKI-40M Update7.8.10 - 7.14.10 Weekly Update

Summary of this Week's Activities:

Wed. 7/7: COMSOL Busbar tutorials; began stack design; began base; Viton rubber research

Thurs. 7/8: Completed Viton rubber research; updated materials; finished designing the base layer

Fri. 7/9: Research model coupling papers; extensive eLog entry about base design and troubleshooting

Sun. 7/11: Played around with Busbar to find first eigenfrequency; continued crashing COMSOL

Mon. 7/12: Intrusions in COMSOL eLog tutorial entry; research eigenfrequency analysis; successfully got first eigenmode of rectangular bar

Tues. 7/13: Updated Poisson ratio of Viton and subsequently succeeded in running eigenfrequency tests on base stack layer. Systematic Perturbation Tests were documented in the most recent elog entry. Discussed results with Rana and decided this didn't make sense. Analytical study required.

Wed. 7/14: Went over to machine shop to experimentally extrapolate spring constant of Viton. Calculations to be done in the afternoon.

  3220   Wed Jul 14 16:39:06 2010 JenneUpdateDAQDebugging Guralp and reboots

[Joe, Jenne]

Joe got on the phone with Alex, and Alex's magic Alex intuition told him to ask about the RFM switch.  The C0DAQ_CTRL's overload light was orangeAlex suggested hitting the reset button on that RFM switch, which we did. That fixed everything -> c0dcu1 came back, as did the frame builder.  Rana had pointed out earlier that we could have brought back all of the other front ends, and enabled the damping of the optics even though the FB was still down.  It's okay to leave the front ends & watchdogs on, and just reboot the FB, AWG, and DAQ_CTRL computers if that is necessary.

Anyhow, once the FB was back online, we got around to bringing back all of the front ends (as usual, except for the ones which are unplugged because they're in the middle of being upgraded).  Everything is back online now.

After all of this craziness, all of the Guralp channels are working happily again. It is still unknown why they starting being digital zero, but they're back again. Maybe I should have rebooted the frame builder in addition to c0dcu1 last night?

 

Quote:

This is regards to zero signal being reported by the channels C1:PEM-SEIS_GUR1_X, C1:PEM-SEIS_GUR1_Y, and C1:PEM-SEIS_GUR1_Z.

I briefly swapped Guralp 1 EW and Guralp 2 EW to confirm to myself that it was not on the gurlap end (although the fact that its digital zero is highly indicative a digital realm problem).  I then unplugged the 17-32, and then 1-16 channel connections to the 110B.  I saw floating noise on the GUR2 channels, but still digital zero on the GUR1 channels, which means its not the BNC break out box.

There was a spare 110B, unconnected in the crate, so to do a quick test of the 110B, I turned off the crate and swapped the 110Bs, after copying the switch configuration of the first 110B to the second one.  The original 110B was labeled ADC 1, while the second 110B was labeled ADC 0.  The switches were identical except for the ones closest to the Dsub connectors on the front.  All those switches in that set were to the right, when looking down at the switches and the Dsub connectors pointing towards yourself.

Unfortunately, the c0duc1 never seemed to come up with the new 110B (ADC 0).  So we put the original 110B back.  And turned the crate back on. 

The fb then didn't seem to come back quite right.  We tried rebooting fb40m it, but its still red with status 1.  c0daqctrl is green, but c0dcu1 is red, although I'm not positive if thats due to fb40m being in a strange state.  Jenne tried a telnet in to port 8087 and shutdown, but that didn't seem to help.  At this point, we're going to contact Alex when he gets in around 12:30.

 

 

  3221   Wed Jul 14 18:09:50 2010 josephb, razibUpdatePhase CameraSome cleanup behind 1Y2 rack of phasecamera electronics

We made an attempt at cleaning up the phase camera setup electronics.

We have moved a portion of the electronics onto the SP table (specifically the mixer, splitters, amplifiers, and associated power).  We put away a large number of cables which were unneeded, both BNC and power cables. The Innolight Mephisto power supply and one signal generator are still behind 1Y2 on top of a non-functioning VME crate.  The second VME crate was put along the south arm where two other VME crates already were.  We placed a fair number of BNC cables and power cords back on their cable racks or approriate storage space, so the rats nests of cables has been reduced.

We moved one power strip from plugging in beyind 1Y1, to the far side of the SP table (closer to the 1Y3 rack), and also found and plugged in another power strip (also on the far side of the SP table) and placed this underneath the SP table to be able to power the signal generator and Innolight Mephisto laser (its not plugged in currently, but we'd like to do so next week).

 

  3224   Wed Jul 14 19:36:17 2010 GopalUpdateOptic StacksExperimental Confirmation of COMSOL Analysis

I experimentally determined the spring constant of a single Vitol spring in order to obtain a rough estimate for the natural frequency of single-stack oscillation.

The procedure basically involved stacking metal bars of known mass onto the Vitol and using a caliper to measure deviations from the equilibrium length.

The plot below shows that, for small compressions, the response is linear to an R-squared of 0.98.

Untitled.png

The experimental spring constant came out to be about 270 lb/ft, or 3900 N/m.

Previous documents have listed that the top stack takes on a load of approximately 43 kg. per individual spring. A bit of calculation yields an experimental resonant frequency of 9.5 Hz.

Compared with the theoretical COMSOL first harmonic of about 7.5 Hz, there is a reasonable amount of error. Of course, I used this calculation as a simple ballpark estimate: errors from misplacement onto the Viton were minimized through use of a level, but were still inevitable on the mm scale. Since the two methods yield answers with the same order of magnitude, we are ready to move forward and build the remaining layers of the stack.

  3226   Thu Jul 15 11:58:50 2010 josephbUpdateComputersAdded channel to ADCU_PEM (C0DCU1)

I modified the C1ADCU_PEM.ini file in /cvs/cds/caltech/chans/daq/ (after making a backup), and added a temporary channel called C1:PEM-TEMP_9, the 9 corresponding to the labeled 9 channel on the front of the BNC breakout in the 1Y7 rack.  The chnnum it was set to is 15008 (it was commented out and called C1:PEM-PETER_FE).  I also set the data rate to 2048.

I then did telnet fb40m 8087, and shutdown, and also hit the blue reconfig button on the DAQ status screen for the C0DCU1 machine.  The framebuilder came back up.  I confirmed the temporary channel, as well as the Guralp channels were still working from C0DCU1.

We have strung a cable in the cable trays from the SP table to the 1Y7 rack, which has been labeled as "Phasecam PD".  This will be used to record the output of an additional photodiode.

 

  3230   Thu Jul 15 16:57:31 2010 josephb, kiwamuUpdateCDSThe temporarily named c1scx machine getting ready to connect to IO chassis

This is the machine which will be the new x end front end machine.  Its IP is 192.168.113.86. 

We changed the root and controls passwords to the usual.  We have modified the controls user group to be 1001, by using "usermod -u 1001 controls" (we had to use the non-rtl kernel to get that command to work).

We changed /etc/fstab to point to /cvs/cds on Linux rather than some downs machine.  We added a link to /cvs/cds/rtcds in the local /opt directory.

We modified the /etc/rc.d/rc.local file to no longer run /opt/open-mx/sbin/omx_init start, /cvs/cds/geo/target/fb/mx_stream -d scipe12:0 om1, and /cvs/cds/geo/target/fb/mx_stream -d scipe12:0 -e2 -r2 om2.  We modified the /usr/bin/setup_shmem.rtl to run only c1x00 c1scx and c1spx.

I also commented out a line0 "/bin/rm -f /rtl*"

 

  3231   Thu Jul 15 19:13:03 2010 ranaUpdateIOOMC_F check

Sometimes I like to plot the spectrum of MC_F. Its a good diagnosis of whether something is wrong.

The red trace is noisier than the blue reference. What is the cause of this?

Attachment 1: a.png
a.png
  3234   Fri Jul 16 12:36:00 2010 Katharine, SharmilaUpdateeloglevitation

After last night's challenge (or inspiration), we levitated our magnet this morning.  Since the nice Olympus camera is not currently in the 40m, we had to use my less stellar camera, but despite the poor video quality you can still see the magnet returning to its stable equilibrium position.  Once we recover the better camera, we will post new videos.  Also, we haven't yet figured out how to put videos in line in the elog entry, so here are the youtube links:

 

levitation 1

levitation 2

 

We adjusted the gain on coil 1 so that the resistance from the pots was 57.1k (maximum gain of 101.2,).

currents from power supply, pre-levitation: 0.08 A and 0.34 A

post levitation: 0.08 A and 0.11 A


note: we're not sure why changing the gain on coil 3 changes the current through the power supply, so we'd like to investigate that next.

Attachment 1: CIMG0649.AVI
  3235   Fri Jul 16 13:05:48 2010 Kyung-haUpdateSUSLate update for 7/13 Tue (Tip Tilts)

[Jenne & Kyung-ha]

We suspended the mirror to one of the main frame with the ECD backplane we finished before. The hard task was to find the right balance for the mirror so that 1) it won't be tilted and 2) it'll be in the right position for the ECD backplanes so that the magnets attached to the mirror holder would be in the very center of each ECD holes. We used optical lever laser (red He/Ne) to check the balance of the mirror. We tried to use the jig for the mirror holder clamps but because of the size difference, we couldn't use it at all. (Since the magnets are very heavy, we thought the wire being not perfectly centered might work better. However, the jig dimension was way too different that the wire ended up in the middle of one of the holes.) Since there was no other clever way to attach the wire in the right position, we just tried to be as center/accurate as possible. After attaching wire to that mirror holder clamps, we hanged it to the frame. Again, we couldn't find any other accurate way to find the center so we held the wire and tried to adjust the mirror height as accurate as possible so that it can be in the right position in respect to ECD backplane and not be tilted at the same time. However, when we hanged the mirror, it was still tilted.. So we adjusted the mirror tilt using the mirror holder clamps. Since the holes on the clamps were ellipse shapes, we could adjust the position of the clamps a little bit. When we adjust the clamps, we started to tighten the screws when the mirror is NOT in the perfect position since the tightening up part changes the mirror angle anyways. Luckily, when we tightened up the last screw, the mirror was in the perfect position! After that, we poked the mirror several times to make sure that it comes back to the same place.

Amazingly, we could finish this whole hanging/adjusting process in about 30 mins! :D (Jan said it's because of his amazing moral support. :P Maybe he'll be there to support us everytime we work on the mirrors?)

  3236   Fri Jul 16 15:39:27 2010 nancyUpdateIOOWavefront Sensors- switched off

I tuned the gain of WFS to 0 last night at about 3am.

I turned it back on now.

  3237   Fri Jul 16 15:57:19 2010 josephb,kiwamuUpdateComputersNew X end FE and IO chassis work

We finished setting up the new X end front end machine (still temporarily called c1scx), and attached it to its IO chassis.  We're preparing for a test tomorrow, where we redirect the Limo breakout box to the new front end and IO chassis, so Kiwamu can test getting some green locking channels into his controls model.

We strung a pair of blue fibers from the timing master to the new X end (and labeled them), so we have a timing signal for the IO chassis.  I also labeled the orange fiber Alex had repurposed from the RFM to timing for the new Y end when I noticed he had not actually labelled it at the timing master.

  3238   Fri Jul 16 16:07:14 2010 josephbUpdateComputersPossible solution for the last ADC

After talking with Jenne, I realized the ADC card in the c1ass machine was currently going unused.  As we are short an ADC card, a possible solution is to press that card into service.  Unfortunately, its currently on a PMC to PCI adapter, rather than PMC to PCIe adapter.  The two options I have are to try to find a different adapter board (I was handed 3 for RFM cards, so its possible there's another spare over in downs - unfortunately I missed Jay when I went over at 2:30 to check).  The other option is put it directly into a computer, the only option being megatron, as the other machines don't have full length PCI slot. 

I'm still waiting to hear back from Alex (who is in Germany for the next 10 days) whether I can connect both in the computer as well as with the IO chassis.

So to that end, I briefly turned off the c1ass machine, and pulled the card.  I then turned it back on, restarted all the code as per the wiki instructions, and had Jenne go over how it looked with me, to make sure everything was ok.

There is something odd with some of the channels reading 1e20 from the RFM network.  I believe this is related to those particular channels not being refreshed by their source (which is other suspension front end machines), so its just sitting at a default until the channel value actually changes.

 

 

  3240   Fri Jul 16 20:25:52 2010 MeganUpdatePSLReference Cavity Insulation

Rana and I

1) took the temperature sensors off the reference cavity;

2) wrapped copper foil around the cavity (during which I learned it is REALLY easy to cut hands with the foil);

3) wrapped electrical tape around the power terminals of the temperature sensors (color-coded, too! Red for the out of loop sensor, Blue for the first one, Brown for the second, Gray for the third, and Violet for the fourth. Yes, we went with an alphabetical coding system, excluding the out of loop sensor);

4) re-wrapped the thermal blanket heater;

5) covered the ends of the cavities with copper, ensuring that the beam can enter and exit;

6) took pretty pictures for your enjoyment!

We will see if this helps the temperature stabilization of the reference cavity.

 

DSC_2271.JPG

The end of the reference cavity, with a lovely square around the beam.

 

DSC_2266.JPG

The entire, well-wrapped reference cavity!

  3241   Fri Jul 16 23:53:27 2010 RanaUpdatePSLReference Cavity Insulation

From the trend, it seems that the Reference Cavity's temperature servo is working fine with the new copper foil. I was unable to find the insulating foam anywhere, but that's OK. We'll just get Frank to make us a new insulation with his special yellow stuff.

The copper foil that Steve got is just the right thickness for making it easy to form around the vacuum can, but we just have to have the patience to wrap ~5-10 more layers on there. We also have to get a new heater jacket; this one barely fits around the outside of the copper wrap. The one we have now seems to have a good heating wire pattern, but I don't know where we can buy these.

I also turned the HEPA's Variac back down to the nominal value of 20. Please remember to turn it back up to 100 before working on the PSL.

  3243   Mon Jul 19 13:51:09 2010 kiwamuUpdateCDSTIming card at X end

[Joe, Kiwamu]

 

 The timing slave in the IO chassis on the new X end was not working with symptoms of no front "OK" green light, no "PPS" light, 3.3V testpoint not working and  ERROR testpoint bouncing between 5-6V.

We took out the timing slave from the X end IO chassis put in to the new Y end IO chassis .

It worked perfectly there. We took the working one from Y end put in the X end IO chassis.

We slowly added cables. First we added power , it worked fine and we saw green "OK" light. Then we added 1PPS signal by a fiber and it also worked.

We turned everything off and then we added 40pin IPC cable from the chassis and infiniband cable from the  computer.

When we turned ON it we didn't see the green light.

This means something in the computer configuration might be wrong not in the timing card, we now are trying to make contact with Alex.

We are comparing the setup of the C1SCX  machine and the working C1ISCEX machine.

  3244   Mon Jul 19 14:14:03 2010 nancy, kojiUpdateIOOQPD Response Transfer Function

Friday night myself and Koji measured the Transfer function of the QPD circuit at MC2 side using a chopper . Following was our procedure :

 

We connected some wires at the input and output of the filter circuit to one of the segment of teh QPD. - seg 1.

A laser light was shined on to the QPD, it was pulsed using a chopper. The frequency of rotation of the chopper was varied.

These wires were then fed to the spectum analyser , and a transfer funstion was observed, It was nearly a low pass filter

The chopper frequency was then made variable by giving the chopper a signal from the spectrum analyser. This signal just swiped a large range of the rpm of the chopper.

Now the input signal looked like a sine wave of varying frequency. the transfer function looked like a perfect LPF, with a small SNR.

Attaching the plot of the TF in the next e-log (this one is on windows and can't access /cvs/cds)

 

  3245   Mon Jul 19 14:16:01 2010 nancy, kojiUpdateIOOQPD Response Transfer Function

Quote:

Friday night myself and Koji measured the Transfer function of the QPD circuit at MC2 side using a chopper . Following was our procedure :

 

We connected some wires at the input and output of teh filter circuit to one of the segment of teh QPD. - seg 2.

A laser light was shined on to  the QPD, it was pulsed using a chopper. The frequency of rotation of teh chopper was varied.

These wires were then fed to the spectum analyser , and a transfer funstion was observed, It was nearly a low pass filter

The chopper frequency was then made variable by giving the chopper a signal from teh spectrum analyser. This signal just swiped a large range of the rpm of the chopper.

Now the input signal looked like a sine wave of varying frequency. the transfer functino looked like a perfect LPF, with a small SNR.

Attaching the plot of the TF in the next e-log (this one is on windows and cant access /cvs/cds)

 

 QPDTF2.png

  3246   Mon Jul 19 16:11:17 2010 GopalUpdateOptic StacksEigenfrequency Analysis of Full Stack

Expanding on the single-layer model, I added the second, third, and fourth layers to the stack in COMSOL. Eigenfrequency analysis run times increased exponentially as the model multiplied in complexity. The following images document the some of the important eigenfrequencies:


First Eigenmode: y-translational, 3.34 Hz:

Eigenfrequency_1_4Stacks.png


Second Eigenmode: x-translational, 3.39 Hz:

Eigenfrequency_2_4Stacks.png


Third Eigenmode: z-rotational, 3.88 Hz:

 Eigenfrequency_3_4Stacks.png


Sixth Eigenmode: z-translational, 8.55 Hz:

Eigenfrequency_6_4Stacks.png


As expected, the eigenfrequencies are generally lower, but still in the same range, as the single-layer model, because of greater mass but constant weight-per-spring distribution.

Next Steps:

1) Extend a single stack to the full stack system, which consists of three stacks like this. Perform similar eigenmode analysis.

2) Analyze the mirror suspension system and incorporate a similar pendulum on the top plate.

3) Make transfer function measurements between seismic and mirror motions.

  3249   Tue Jul 20 11:49:31 2010 JenneUpdateSUSSome Suspensions not damping

[Jenne, Koji]

I moseyed into the control room this morning, to find ITMX and ITMY both with their watchdogs tripped.  ITMY (new convention) wouldn't damp.  Koji discovered that there was a sign flip in 2 of the sensors.  A set of reboots of c1susvme1&2 fixed the problem. 

A side note:  For the ETMs, the OSEM sensor readouts are gigantic (~20,000), whereas for the similar channels on all other optics, the readouts are on the order of 1.  After some looking around, it seems that this is just the way things have been (for at least 100 days), and the filters in the SUSPOS and other SUS filter banks have a high pass filter to take care of this.  It's weird, but it seems to be the way it is, and the ETMs damp, so it's all good.

  3250   Tue Jul 20 11:55:15 2010 Sharmila,KatharineUpdateelogMaglev

We plotted the transfer functions for the maglev control circuit and compared them with the plots from the spectrum
analyzer. We were stuck for sometime because

1) we had wrongly entered the value of one of the resistors which was off by a factor of 2000.
2) The plots were not done in right units. So we couldn't figure out differences quite well.

The two plots are shown below. We are still off by a factor of 3 which we'll figure out soon.

P7140292.JPG

  3253   Tue Jul 20 18:29:43 2010 nancyUpdateIOOQPD installed behind the MC2

 

Yesterday I installed teh QPD on the table behind MC2, and observed teh signal on it.

The MC_leak is directed to it by a steering mirror.

I used the A2L_MC2 script to minimise  teh pitch and yaw gains, and estimated teh spot position on teh MC2 using that.

This spot position was aligned to the center of teh QPD.

In the night while before taking measurements, I decided to turn off the Wavefront Sensor Servos, but just after that, the MC alignment went very bad, and I could not align it in the next 2 hours.

For some reason, the MC was really mad the whole day yesterday, and was getting misaligned again and again, even when the WFS feedback was on.

 

The table also had another IR laser in it, which I and Koji switched off.

 

I will continue measuring once we pump down again.

For now, I am analysing teh QPD circuit Transfer Function.