  5441   Fri Sep 16 21:36:25 2011 KeikoUpdateLSCPOY11 and POY55 were added

 New channels, POP55 and POY11 are connected to the rack and now available on the data system.

POX11 I is not working. I didn't investigate what was wrong. Please make sure when you come to need POX11.

The orthogonalities of POY11 and POP55 were measured and already adjusted. The results are below:

POY11

ABS = 0.973633 

PHASE = 92.086483 [deg]

ezcawrite C1:LSC-POY11_Q_GAIN 1.027081 && ezcawrite C1:LSC-POY11_PHASE_D 92.086483

POP55

ABS =  1.02680579

PHASE =  88.5246 [deg]

ezcawrite C1:LSC-POP55_Q_GAIN 0.973894 && ezcawrite C1:LSC-POP55_PHASE_D 88.524609

 

 


  5445   Sat Sep 17 01:53:41 2011 KeikoUpdateLSCPOY and POP beams clipped

 Keiko, Paul, Kiwamu

We found that POP beam is clipped by the steering mirrors inside the tank. POY beam is also likely to be clipped inside. Also the hight of POY beam is too high (about 5 cm higher than the normal paths) at the first lens. These imply the input pointing is bad.

  5464   Mon Sep 19 16:44:16 2011 KeikoHowToLSCProcedure for the demodulation board check

 Here I note the procedure for the demodulation board orthogonality check for the future reference.

1. prepare two function generators and make sure I an Q demodulation signals go to the data acquisition system.

2. sync the two generators

3. drive the function generator at the modulation frequency and connect to the LO input on the demod board

4. drive the other function generator at the modulation frequency + 50Hz  the RF in

5. run "orthogonality.py"  from a control computer scripts/demphase directory. It returns the amplitude and phase information for I and Q signals. If necessary, compensate the amplitude and phase by the command that  "orthogonality.py" returns.

 

If you want to check in the frequency domain (optional):

1. 2. 3 are the same as above.

4. drive the function generator at the LO frequency + sweep the frequency, for example from 1Hz to 1kHz, 50ms sweep time. You can do it by the function generator carrier frequency sweep option.

5. While sweeping the LO frequency, run "orthogonality.py"

6. The resulting plot from "orthogonality.py" will show the transfer function from the RF to demodulated signal. The data is saved in "dataout.txt" in the same directory.

  5472   Mon Sep 19 23:19:40 2011 KeikoUpdateIOOAM modulation mistery

 Keiko, Anamaria

We started to investigate the AM modulation mistery again. Checking just after the EOM, there are AM modulation about -45dBm. Even if we adjust the HWP just before the EOM, AM components grow up in 5 mins. This is the same situation as before. Only the difference from before is that we don't have PBS and HWP between the EOM and the monitor PD. So we have a simpler setup this time.

We will try to align the pockells cell alignment tomorrow daytime, as it may be a problem when the crystal and the beam are not well parallel. This adjustment has been done before and it didn't improve AM level at that time.

  5474   Tue Sep 20 03:02:23 2011 KeikoUpdateLSClocking activity tonight

 Keiko, Anamaria, Koji

We were not able to establish the stable DRMI tonight. We could lock MICH and PRCL quite OK, and lock the three degrees of freedom at somewhere strange for several seconds quite easily, but the proper DRMI lock was not obtained.

When MICH and PRC are locked to the carrier, REFL DC PD reading dropps from ~3000 counts to 2600~2700 counts as REFL beam is absorbed to PRC. We'll try to lock PRC to sidebands - but flipping gain sign didn't work today, although it worked a few days ago. 

POP beam (monitor) is useful to align PRM.

  5483   Tue Sep 20 16:31:24 2011 KeikoUpdateIOOSmall modulation depth

 Modulation resonator box is removed and the modulation depth is small right now.

I have broke the BNC connector on the modulation resonator box. The connector was attached by the screw inside very loosely and when we connect and disconnect the BNC cables from outside, extra force was applied to the cable inside and it was broke. It is being fix by Kiwamu and will be back in a bit.

 

 

 

 

  5484   Tue Sep 20 16:38:25 2011 KeikoUpdateIOOSmall modulation depth

Resonator box and the modulations are back now. But the modulation depth seems to be a bit smaller than yesterday, looking at the optical spectrum analyser.

 

Quote:

 Modulation resonator box is removed and the modulation depth is small right now.

I have broke the BNC connector on the modulation resonator box. The connector was attached by the screw inside very loosely and when we connect and disconnect the BNC cables from outside, extra force was applied to the cable inside and it was broke. It is being fix by Kiwamu and will be back in a bit

 

  5491   Tue Sep 20 23:01:37 2011 KeikoUpdateIOOAM modulation mistery

Keiko, Suresh

AM modulations are still there ... the mechanical design for the stages, RF cables, and connections are not good and affecting the alignment.

I write the activity in the time series this time - Because we suspect the slight EOM misalignment to the beam produces the unwanted AM sidebands, we tried to align the EOM as much as possible. First I aligned the EOM tilt aligner so that the maximum power goes through. I found that about 5% power was dumped by EOM. After adjusting the alignment, the AM modulation seemed be much better and stable, however, it came up after about 20 mins. They grew up up to about -40dBm, while the noise floor is -60 dBm (when AM is minimised, with DC power of 8V by PDA225 photodetector).

We changed the EOM stage (below the tilt aligner) from a small plate to a large plate, so that the EOM base can be more stable. The EOM stands on the pile of several black plate. There was a gap below the tilt aligner because of a small plate.  So we swapped the small plate to large plate to eliminate the springly gap. However it didn't make any difference - it is the current status and there is still AM modulations right now.

During above activities, we leaned that the main cause of the EOM misalignment may be the RF cables and the resonator box connected to the EOM. They are connected to the EOM by an SMA adaptor, not any soft cables. It is very likely applying some  torc force to the EOM box. The resonator box is almost hunging from the EOM case and just your slight touch changes EOM alinment quite a bit and AM mod becomes large. 

I will replace the SMA connector between the resonator box and EOM to be a soft cable, so that the box doesn't hung from EOM tomorrow. Also, I will measure the AM mod depth so that we compare with the PM mod depth.

 

Quote:

 Keiko, Anamaria

We started to investigate the AM modulation mistery again. Checking just after the EOM, there are AM modulation about -45dBm. Even if we adjust the HWP just before the EOM, AM components grow up in 5 mins. This is the same situation as before. Only the difference from before is that we don't have PBS and HWP between the EOM and the monitor PD. So we have a simpler setup this time.

We will try to align the pockells cell alignment tomorrow daytime, as it may be a problem when the crystal and the beam are not well parallel. This adjustment has been done before and it didn't improve AM level at that time.

 

  5495   Wed Sep 21 02:49:39 2011 KeikoSummaryLSCLSC matrices

I created 3 kinds of LSC matrices, PRMI condition with carrier resonant in PRC, PRMI condition with SB resonant in PRC, and DRMI with SB resonant in PRC. The matrices are with AS55 and REFL11 which are used for locking right now. The signal numbers are written in log10, and the dem phases are shown in degrees.

From CR reso PRMI to SB reso PRMI, demodulation phases change  ----

 

PRMI - Carrier resonant in PRC

 

            PRCL      MICH  SRCL

REFL11 7.7079 2.9578 0
REFL33 5.2054 3.2161 0
REFL55 7.7082 2.9584 0
REFL165 3.9294 2.5317 0
AS11 1.0324 3.5589 0
AS33 1.0286 1.6028 0
AS55 1.1708 4.2588 0
AS165 1.1241 0.9352 0
POP11 2.8015 -1.3331 0
POP33 0.2989 -1.6806 0
POP55 2.8017 -0.6493 0
POP165 -0.9769 -2.3708 0
POX11 3.7954 -0.3363 0
POX33 1.293 -0.7058 0
POX55 3.796 0.355 0
POX165 0.0187 -1.3837 0
       
Dem Phase      
REFL11 3 179 0
REFL33 165 -172 0
REFL55 13 170 0
REFL165 86 177 0
AS11 -32 73 0
AS33 176 -72 0
AS55 -41 12 0
AS165 -7 146 0
POP11 -11 -116 0
POP33 124 147 0
POP55 -54 -146 0
POP165 -117 -25 0
POX11 -87 15 0
POX33 -105 -80 0
POX55 -76 16 0
POX165 180 -91 0

PRMI - SB resonant in PRC

SB reso PRMI    
  PRCL MICH SRCL
REFL11 7.6809 5.2777 0
REFL33 5.2465 3.1565 0
REFL55 7.2937 5.589 0
REFL165 4.3892 2.6857 0
AS11 1.3123 3.545 0
AS33 0.9331 1.6022 0
AS55 1.7425 4.0514 0
AS165 1.5838 1.1344 0
POP11 2.7745 0.3791 0
POP33 0.3401 -1.7392 0
POP55 2.3872 0.6904 0
POP165 -0.5171 -2.2279 0
POX11 3.7684 1.3574 0
POX33 1.3341 -0.7664 0
POX55 3.3815 1.6688 0
POX165 0.4785 -1.2163 0
       
Dem Phase
     
REFL11 155 -115 0
REFL33 -8 3 0
REFL55 91 -178 0
REFL165 -62 28 0
AS11 109 62 0
AS33 -39 99 0
AS55 13 -38 0
AS165 -155 168 0
POP11 141 -128 0
POP33 -48 -38 0
POP55 24 115 0
POP165 95 -176 0
POX11 65 155 0
POX33 83 95 0
POX55 2 92 0
POX165 32 123 0

DRMI - SB resonant in PRC

REFL11 7.6811 5.0417 4.2237 
REFL33 5.2751 4.1144 3.7766
REFL55 7.2345 7.0288 6.6801
REFL165 4.3337 4.1266 3.7775
AS11 1.1209 3.512 0.9248
AS33 0.9159 1.6323 0.7971
AS55 2.6425 5.3915 2.5519
AS165 2.6423 2.4881 2.3272
POP11 2.7747 0.1435 -0.6846
POP33 0.3687 -0.7849 -1.122
POP55 2.3244 2.1302 1.7815
POP165 -0.5833 -0.8 -1.1548
POX11 3.7676 3.261 0.8086
POX33 1.3896 0.2372 0.2333
POX55 3.4619 3.0097 3.1326
POX165 0.782 0.6668 0.4357
                        
Dem Phase
     
REFL11 154 -16 4
REFL33 -5 12 51
REFL55 129 -166 -123
REFL165 -23 40 83
AS11 132 79 69
AS33 -92 -127 -83
AS55 -33 -55 -5
AS165 154 179 -144
POP11 141 -29 -9
POP33 -46 -27 12
POP55 62 127 170
POP165 135 -161 -117
POX11 64 -102 -83
POX33 85 143 118
POX55 57 103 124
POX165 99 155 -164

 

 

  5502   Wed Sep 21 16:44:18 2011 KeikoUpdateIOOAM modulation mistery

AM modulation depths are found to be 50 times smaller than PM modulation depths.

m(AM,f1) ~ m(AM, f2) = 0.003 while m(PM, f1)=0.17 and m(PM, f2)=0.19.

Measured values;

* DC power = 5.2V which is assumed to be 0.74mW according to the PDA255 manual.

*AM_f1 and AM_f2 power = -55.9 dBm = 2.5 * 10^(-9) W.

P92101381.jpg

AM f2 power is assumed to be the similar value of f1. I can't measure f2 (55MHz) level properly because the PD (PDA255) is 50MHz bandwidth. From the (P_SB/P_CR) = (m/2) ^2 relation where P_SB and P_CR are the sideband and carrier power, respectively, I estimated the rough the AM modulation depths. Although DC power include the AM SB powers, I assumed that SB powers are enough small and the DC power can be considered as the carrier power, P_CR. The resulting modulation depth is about 0.003.

On the other hand, from the OSA, today's PM mod depths are 0.17 and 0.19 for f1 and f2, respectively. Please note that these numbers contains (small) AM sidebands components too. Comparing with the PM and AM sideband depths, AM sidebands seems to be enough small.

Quote:

Keiko, Suresh

AM modulations are still there ... the mechanical design for the stages, RF cables, and connections are not good and affecting the alignment. 

 

Attachment 1: P9210138.JPG
P9210138.JPG
  5504   Wed Sep 21 18:53:03 2011 KeikoUpdateIOOAM modulation misery

The signal offset due to the AM modulation is estimated by a simulation for PRCL for now. Please see the result below.

Too see how bad or good the AM modulation with 1/50 modulation depths of PM, I ran a simulation. For example I looked at PRCL sweep signal for each channel. I tried the three AM modulation depths, (1) m_AM=0 & m_PM = 0.17 (2) m_AM = 0.003 & m_PM = 0.17 which is the current modulation situation (3) m_AM = 0.17 & m_PM = 0.17 in which AM is the same modulation depth as PM.  For the current status of (2), there are offsets on signals up to 0.002 while the maximum signal amplitude is 0.15. I can't tell how bad it is.... Any suggestions?

 

(1) m_AM=0 & m_PM = 0.17. There is no offset in the signals.

AM0.png

(2) m_AM = 0.003 & m_PM = 0.17. There are offsets on signals up to 0.002 while the maximum signal amplitude is 0.15.

AMratio50.png

(3) m_AM = 0.17 & m_PM = 0.17. There are offsets on signals up to 0.1 while the maximum signal amplitude is 0.2.

AMratio1.png

I will look at MICH and SRCL in the same way. 

Quote:

I'd like to see some details about how to determine that the ratio of 1:50 is small enough for AM:PM.

* What have people achieved in past according to the elogs©  of the measurements?

* What do we expect the effect of 1:50 to be? How much offset does this make in the MICH/PRC/SRC loops? How much offset is too much?

Recall that we are using frontal modulation with a rather small Schnupp Asymmetry...

 

  5512   Thu Sep 22 01:45:41 2011 KeikoUpdateLSCLocking status update

Keiko, Anamaria

Tonight we want to measure the LSC matrix for PRMI and compare the simulation posted last night (#5495).

First. we locked MICH and PRCL, and measured the OLT to see how good the locking is. The following rough swept sine plots are the OLTs for MICH and PRCL. The gain setting was -10 and 0.5 for MICH and PRCL, respectively. Integrators were off. Looking at the measured plots, MICH has about 300 Hz UGF, when the gain is -20, and PRCL has about 300 HZ UGF, too, when the gain is 0.8.

MICH-OLT.pdf

PRCL-OLT.pdf

As these lokings seemed good, so we tried the LSC matrix code written by Anamaria. However it is not working well at this point. When the script add excitations to the exc channels, they kick the optics too much and the lockings are too much disturbed...

Also, we have been trying to lock PRC with the SB resonant, it doesn't work. Looking at the simulated REFL11I (PRCL) signal (you can see it in #5495 too), the CR and SB resonances have the opposite signs... But minus gain never works for PRCL. It only excites the mirror rather than locking.

  5520   Thu Sep 22 17:29:42 2011 KeikoUpdateIOOAM modulation mistery

AM modulation will add offset on SRCL signal as well as PRCL signal. About 2% of the signal amplitude with the current AM level. MICH will not be affected very much.

From #5504, as for the AM modulation I checked the MICH and SRCL signals in addition to the last post for PRCL, to see the AM modulation effect on those signals. On the last post, PRCL (REFL11I) was found to have 0.002 while the maximum signal amplitude is 0.15 we use . Here, I did the same simulation for MICH and SRCL.

As a result, MICH signals are not affected very much. The AM modulation slightly changes signal slopes, but doesn't add offsets apparently. SRCL is affected more, for REFL signals. All the REFL channels get about 0.0015 offsets while the signal ampliture varies up to 0.002. AS55I (currently used for SRCL) has 1e-7 offset for 6e-6 amplitude signal (in the last figure) - which is the same offset ratio comparing with the amplitude in the PRCL case -

 

(1) MICH signals at AS port with AM m=0

AMmod0MICH.png

(2) MICH signals at AS port with AM m=0.003

AMmod1e-1MICH.png

(3) SRCL signals at AS/REFL port with AM m=0

AMmod0SRCL.png 

(3) SRCL signals at AS/REFL port with AM m=0.003

AMmod3e-3SRCL.png

AMmod3e-3SRCL-AS55I.png

 

Quote:

How about changing the x-axis of all these plots into meters or picometers and tell us how wide the PRC resonance is? (something similar to the arm cavity linewidth expression)

Also, there's the question of the relative AM/PM phase. I think you have to try out both I & Q in the sim. I think we expect Q to be the most effected by AM.

 

  5538   Sat Sep 24 09:55:42 2011 KeikoUpdateIOOAM modulation mistery

From the night day before yesterday (Sep 22nd, Thursday night. Sorry for my late update), there are more AM modulations than I measured in the previous post. It is changing a lot, indeed! Looking at the REFL11 I and Q signals on the dataviewer, the signal offset were huge, even after "LSCoffset" script. Probably the modulation index of AM was same order of PM at that time. The level of AM mod index is changing a lot depending on the EOM alingment which is not very stable, and also on the environment such as temperature .

To reduce AM modulations, here I note some suggestions you may want to try :

* Change the SAM connectors between RF resonator and EOM to be a soft but short connector, so that the resonator box doesn't hung from the EOM.

* Change the RF resonator base to be stable posts. Now several black plates are piled to make one base.

* Install a temperature shield

* Also probably you want to change the BNC connector on the RF resonator to be SMA.

* Be careful of the EOM yaw alignment. Pitch seemed to be less sensitive in producing AM than yaw alignment.

 

Quote:

AM modulation will add offset on SRCL signal as well as PRCL signal. About 2% of the signal amplitude with the current AM level. MICH will not be affected very much.

From #5504, as for the AM modulation I checked the MICH and SRCL signals in addition to the last post for PRCL, to see the AM modulation effect on those signals. On the last post, PRCL (REFL11I) was found to have 0.002 while the maximum signal amplitude is 0.15 we use . Here, I did the same simulation for MICH and SRCL.

As a result, MICH signals are not affected very much. The AM modulation slightly changes signal slopes, but doesn't add offsets apparently. SRCL is affected more, for REFL signals. All the REFL channels get about 0.0015 offsets while the signal ampliture varies up to 0.002. AS55I (currently used for SRCL) has 1e-7 offset for 6e-6 amplitude signal (in the last figure) - which is the same offset ratio comparing with the amplitude in the PRCL case -

 

  6358   Mon Mar 5 18:12:00 2012 KeikoUpdateLSCRAM simulation update

 I wrote an RAM simulation script ... it calculates the LSC signal offset and the operation point offset depending on the RAM modulation index.

Configuration : RAM is added on optC1, by the additional Mach-Zehnder ifo before the PRM.

Mar5RAM3.pngMar5RAM2.png

 Both are for PRCL sweep result. Note that REFL33I is always almost zero. Next step: Check the LSC matrix with matrix at the offset operation point.

  6363   Tue Mar 6 15:22:02 2012 KeikoUpdateLSCRAM simulation update

Quote:

 I wrote an RAM simulation script ... it calculates the LSC signal offset and the operation point offset depending on the RAM modulation index.

Configuration : RAM is added on optC1, by the additional Mach-Zehnder ifo before the PRM.

Mar5RAM3.pngMar5RAM2.png

 Both are for PRCL sweep result. Note that REFL33I is always almost zero. Next step: Check the LSC matrix with matrix at the offset operation point.

 On the right figure, you see the non-zero operation points even when RAM mod index = 0. Apparently they come from non-zero loss of the model.  (Each mirror of 50ppm loss was assumed).

  5233   Sun Aug 14 20:04:40 2011 Keiko, Anamaria, Jenne, and KiwamuSummaryLockingcentral part ifo locking plan
GOAL : To lock the central part of ifo

Here is the plan:

Mon - assemble all the cables from PDs and mixers, and check the CDS channels. Prepare the beamsplitters.

Tue - The current paths to REFL11 and REFL55 will be modified to the four paths to REFL11, 33, 55, 165. And the PDs will be placed.
Wed, Thu - during waiting for the ifo available with vacuum, help aligning the POP, POX, POY. In parallel, a simulation to find the PRC length SRC 
length tolerance will be proceeded.

Fri - When the ifo becomes available with vacuum, the sensing signals by 3-f scheme will be obtained with proper demodulation phases.

Sat - Try to lock the central part of the ifo with the new 3-f signals.
  2810   Mon Apr 19 16:31:42 2010 KevinUpdatePSLInnolight 2W Laser

Koji and Kevin

We unpacked the Innolight 2W laser, took an inventory, and scanned the operations manual.

[Edit by KA]

The scanned PDFs are placed on the following wiki page

http://lhocds.ligo-wa.caltech.edu:8000/40m/Upgrade_09/PSL

We will measure the P-I curve, the mode profile, frequency actuator responses, and so on.

  2822   Tue Apr 20 20:15:37 2010 KevinUpdatePSLInnolight 2W Output Power vs Injection Current

Koji and Kevin measured the output power vs injection current for the Innolight 2W laser.

The threshold current is 0.75 A.

 

The following data was taken with the laser crystal temperature at 25.04ºC (dial setting: 0.12).

Injection Current (A) Dial Setting Output Power (mW)
0.000 0.0 1.2
0.744 3.66 1.1
0.753 3.72 4.6
0.851 4.22 102
0.954 4.74 219
1.051 5.22 355
1.151 5.71 512
1.249 6.18 692
1.350 6.64 901
1.451 7.08 1118
1.556 7.52 1352
1.654 7.92 1546
1.761 8.32 1720
1.853 8.67 1855
1.959 9.05 1989
2.098 9.50 2146

 

Attachment 1: PvsI_2W.jpg
PvsI_2W.jpg
  2828   Wed Apr 21 21:56:27 2010 KevinUpdatePSLInnolight 2W Vertical Beam Profile

Koji and Kevin measured the vertical beam profile of the Innolight 2W laser at one point.

This data was taken with the laser crystal temperature at 25.04°C and the injection current at 2.092A.

The distance from the razor blade to the flat black face on the front of the laser was 13.2cm.

The data was fit to the function y(x)=a*erf(sqrt(x)*(x-x0)/w)+b with the following results.

Reduced chi squared = 14.07

x0 = (1.964 +- 0.002) mm

w  = (0.216 +- 0.004) mm

a  = (3.39  +- 0.03) V

b  = (3.46  +- 0.03) V

Attachment 1: bp2.jpg
bp2.jpg
Attachment 2: bp2.dat
razor height (mm)   Voltage (V)
2.75    6.89
2.50    6.90
2.30    6.89
2.25    6.89
2.20    6.75
2.15    6.47
2.13    6.20
2.10    6.05
2.07    5.88
... 17 more lines ...
  2837   Sat Apr 24 15:05:41 2010 KevinUpdatePSL2W Vertical Beam Profile

The vertical beam profile of the Innolight 2W laser was measured at eight points along the axis of the laser.

These measurements were made with the laser crystal temperature at 25.04°C and the injection current at 2.091A. z is the distance from the razor blade to the flat black face of the front of the laser.

The voltage from a photodiode was measured for the razor at a number of heights. Except for the first two points, one scan was made with the razor moving down and a second scan was made with the razor moving up. This data was fit to

y = a*erf(sqrt(2)*(x-x0)/w) + b with the following results:

z(cm) (±0.1cm) w(mm) chi^2/ndf
3.9 0.085 ± 0.006 77.09
6.4 0.130 ± 0.004 12.93
8.8 down 0.145 ± 0.008 66.57
8.8 up 0.147 ± 0.008 18.47
11.6 down 0.194 ± 0.010 64.16
11,6 up 0.214 ± 0.009 27.23
14.2 down 0.177 ± 0.008 49.95
14.2 up 0.183 ± 0.007 29.85
16.6 down 0.205 ± 0.006 18.35
16.2 up 0.203 ± 0.007 17.16
19.2 down 0.225 ± 0.007 18.92
19.2 up 0.238 ± 0.011 25.56
21.7 down 0.292 ± 0.006 11.30
21.7 up 0.307 ± 0.008 11.85

The values for w and its uncertainty were estimated with a weighted average between the two scans for the last six points and all eight points were fit to

w = w0*sqrt(1+(z-z0)2/zR2) with the following results:

chi^2/ndf = 17.88

w0 = (0.07 ± 0.13) mm

z0 = (-27 ± 121) mm

zR = (65 ± 93) mm

It looks like all of the data points were made in the linear region so it is hard to estimate these parameters with reasonable uncertainty.

Attachment 1: vbp.jpg
vbp.jpg
  2846   Mon Apr 26 16:51:37 2010 KevinUpdatePSLre: 2W Vertical Beam Profile

I tried Koji's suggestions for improving the fit to the vertical beam profile; however, I could not improve the uncertainties in the fit parameters.

I started retaking the data today with the same laser settings used last time and noticed that the photodiode was saturating. We were using an ND 4.0 neutral density filter on the photodiode. Koji and I noticed that the coating on the filter was reduced in the center and added an additional ND 0.6 filter to the photodiode. This seemed to fix the photodiode saturation.

I think that the photodiode was also saturating to a lesser extent when I took the last set of data. I will take another vertical beam profile tomorrow.

[Edit by KA: Metallic coating started being evaporated and the ND filters reduced their attenuation. We decided to use absorptive one as the first incident filter, and put a thinner one behind. This looked fine.]

  2851   Tue Apr 27 15:29:16 2010 KevinUpdatePSLre: 2W Vertical Beam Profile

I thought that the micrometer I was using to move the razor through the laser beam was metric; however, it is actually english.

After discovering this mistake, I converted my previous measurements to centimeters and fit the data to

w = sqrt(w0^2+lambda^2*(z-z0)^2/(pi*w0)^2) with the following results:

reduced chi squared = 14.94

z0 = (-4.2 ± 1.9) cm

w0 = (0.013 ± 0.001) cm

Attachment 1: vbp.jpg
vbp.jpg
Attachment 2: vbp_residuals.jpg
vbp_residuals.jpg
  2857   Wed Apr 28 14:22:36 2010 KevinUpdatePSLre: 2W Vertical Beam Profile

I used the Mathematica CurveFit package that we use in Ph6/7 to make the fits for the beam profile data. I wrote two functions that use CurveFit shown in the attachment to make the fits to the error function and square root.

Attachment 1: BeamFit.nb.tar
  2859   Wed Apr 28 16:15:02 2010 KevinUpdatePSLAccelerometer Calibration

Koji, Steve, and Kevin looked into calibrating the Wilcoxon accelerometers. Once calibrated, the accelerometers will be used to monitor the motion of the PSL table.

We want to use the shaker to shake each accelerometer and monitor the motion with an OSEM. We will make a plate to attach an accelerometer to the shaker. A flag will also be mounted on this plate.The OSEM will be mounted on the table next to the shaker and positioned so that the flag can block the LED light as the plate moves up and down. We will then measure the motion of the accelerometer as it is shaken from the OSEM signal. The OSEM signal will be calibrated by keeping the plate and the flag still and moving the OSEM down along the flag a known distance with a micrometer.

  2907   Mon May 10 20:03:22 2010 KevinUpdateGreen LockingGreen Laser Beam Profile

Kiwamu and Kevin measured the beam profile of the green laser by the south arm ETM.

The following measurements were made with 1.984A injection current and 39.65°C laser crystal temperature.

 

Two vertical scans (one up and one down) were taken with a razor blocking light entering a photodiode with the razor 7.2cm from the center of the lens. This data was fit to

b + a*erf(sqrt(2)*(x-x0)/w) with the following results:

scan down: w = (0.908 ± 0.030)mm  chi^2 = 3.8

scan up:      w = (0.853 ± 0.025)mm   chi^2 = 2.9

giving a weighted value of w = (0.876 ± 0.019)mm at this distance.

 

The beam widths for the profile fits were measured with the beam scanner. The widths are measured as the full width at 13.5% of the maximum. Each measurement was averaged over 100 samples. The distance is measured from the back of the lens mount to the front face of the beam scanner.

distance (cm) vertical w (µm) horizontal w (µm)
3.2 ± 0.1 1231 ± 8 1186 ± 7
4.7 ± 0.1 1400 ± 4 1363 ± 6
7.4 ± 0.1 1656 ± 5 1625 ± 9
9.6 ± 0.1 1910 ± 10 1863 ± 9
12.5 ± 0.1 2197 ± 8 2176 ± 8
14.6 ± 0.1 2450 ± 12 2416 ± 10
17.5 ± 0.1 2717 ± 12 2694 ± 14
20.0 ± 0.1 2973 ± 16 2959 ± 8
22.4 ± 0.1 3234 ± 12 3193 ± 14

This data was fit to w = sqrt(w0^2+lambda^2*(x-x0)^2/(pi*w0)^2) with lambda = 532nm with the following results:

For the vertical beam profile:

reduced chi^2 = 3.29

x0 = (-87   ± 1)    mm

w0 = (16.30 ± 0.14) µm

For the horizontal beam profile:

reduced chi^2 = 2.01

x0 = (-82   ± 1)    mm

w0 = (16.12 ± 0.10) µm

Note: These fits were done with the beam diameter instead of the beam radius. The correct fits to the beam radius are here: http://nodus.ligo.caltech.edu:8080/40m/2912

Attachment 1: vbp.jpg
vbp.jpg
Attachment 2: vbp_residuals.jpg
vbp_residuals.jpg
Attachment 3: hbp.jpg
hbp.jpg
Attachment 4: hbp_residuals.jpg
hbp_residuals.jpg
  2912   Tue May 11 17:02:43 2010 KevinUpdateGreen LockingGreen Laser Beam Profile

 

Quote:

Hey, what a quick work!

But, wait...

1) The radius of the beam was measured by the razor blade.

2) The diameter of the beam (13.5% full-width) at each point was measured by Beam Scan. The one at z=~7cm was consistent with 1)

3) The data 2) was fitted by a function w = sqrt(w0^2+lambda^2*(x-x0)^2/(pi*w0)^2). This is defined for the radius, isn't it?

So the fitting must be recalculated with correct radius.
Make sure that you always use radius and write with a explicit word "radius" in the record.

I recalculated the fits using the radius of the beam instead of the diameter of the beam at 13.5% full-width with the following results:

For the vertical beam profile:

reduced chi^2 = 3.25

x0 = (-86 ± 1)mm

w0 = (46.01 ± 0.38)µm

For the horizontal beam profile:

reduced chi^2 = 2.05

x0 = (-81 ± 1)mm

w0 = (45.50 ± 0.28)µm

Attachment 1: vbp.jpg
vbp.jpg
Attachment 2: vbp_residuals.jpg
vbp_residuals.jpg
Attachment 3: hbp.jpg
hbp.jpg
Attachment 4: hbp_residuals.jpg
hbp_residuals.jpg
  2933   Fri May 14 16:14:37 2010 KevinUpdateGreen LockingGreen Laser Beam Profile

Quote:

Strange. I thought the new result became twice of the first result. i.e. w0=32um or so.

Can you explain why the waist raidus is estimated to be three times of the last one?
Can you explain why the measured radius @~70mm is not 0.8mm, which you told us last time,
but is 0.6mm?

The measurements have been done at the outside of the Rayleigh range.
This means that the waist size is derived from the divergence angle

theta = lambda / (pi w0)

At the beginning you used diameter instead of radius. This means you used twice larger theta to determine w0.
So if that mistake is corrected, the result for w0 should be just twice of the previous wrong fit.

 

 

I was off by a factor of sqrt(2). The correct fit parameters are

for the vertical beam profile:

reduced chi^2 = 3.28

x0 = (-87 ± 1) mm

w0 = (32.59 ± 27) µm

for the horizontal beam profile

reduced chi^2 = 2.02

x0 = (-82 ± 1) mm

w0 = (32.23 ± 20) µm

In the following plots * denotes vertical data points and + denotes horizontal data points. The blue curve is the fit to the vertical data and the purple curve is the fit to the horizontal data.

Attachment 1: profile.png
profile.png
Attachment 2: residuals.png
residuals.png
  2976   Mon May 24 16:34:22 2010 KevinUpdatePSLND Filters for 2W Beam Profile

I tried to measure the beam profile of the 2W laser today but ran into problems with the ND filters. With the measurements I made a few weeks ago, I used a reflective ND 4.0 filter on the PD. The PD started to saturate and Koji and I noticed that a lot of the metallic coating on the filter had been burnt off. Koji told me to use an absorptive ND 4.0 filter in front of a reflective ND 0.6 filter. I tried this today but noticed that a few holes were being burned into the absorptive filter and that the coating on the reflective filter behind it was also being burned off in spots. I don't think we wanted to use a polarizing beam splitter to reduce the power before the PD but I didn't want to ruin any more filters.

  2984   Tue May 25 17:04:37 2010 KevinUpdate Beam Profile After Mode Cleaner

I fit the data from the beam profile that Jenne measured on 5/21/2010. The distances are measured from halfway between MC1 and MC3 to the beam scanner. The fits give the following where w0 is the waist size and z0 is the distance from the waist to halfway between MC1 and MC3.

For the horizontal profile:

reduced chi^2 = 0.88

z0 = (1 ± 29) mm

w0 = (1.51 ± 0.01) mm

For the vertical profile:

reduced chi^2 = 0.94

z0 = (673 ± 28) mm

w0 = (1.59 ± 0.01) mm

I calculated the radius of curvature of MC2 using these values of w0:

horizontal: (16.89 ± 0.06) m

vertical:   (17.66 ± 0.07) m

For this calculation, I used the value of (13.546 ± .0005) m for the length of the mode cleaner measured on 6/10/2009. The specification for the radius of curvature of MC2 is (18.4 ± 0.1) m.

In the following plots, the blue curve is the fit to the vertical beam radius, the purple curve is the fit to the horizontal beam radius, * denotes a data point from the vertical data, and + denotes a data point from the horizontal data.

Attachment 1: mcfit.png
mcfit.png
Attachment 2: mcerrors.png
mcerrors.png
  2986   Tue May 25 17:22:56 2010 KevinUpdateIOOBeam Profile After Mode Cleaner

Quote:

Very nice as usual. Can you add the curve to show the ideal mode of the MC on the profile plot?

Quote:

I fit the data from the beam profile that Jenne measured on 5/21/2010. The distances are measured from halfway between MC1 and MC3 to the beam scanner. The fits give the following where w0 is the waist size and z0 is the distance from the waist to halfway between MC1 and MC3.

For the horizontal profile:

reduced chi^2 = 0.88

z0 = (1 ± 29) mm

w0 = (1.51 ± 0.01) mm

For the vertical profile:

reduced chi^2 = 0.94

z0 = (673 ± 28) mm

w0 = (1.59 ± 0.01) mm

I calculated the radius of curvature of MC2 using these values of w0:

horizontal: (16.89 ± 0.06) m

vertical:   (17.66 ± 0.07) m

For this calculation, I used the value of (13.546 ± .0005) m for the length of the mode cleaner measured on 6/10/2009. The specification for the radius of curvature of MC2 is (18.4 ± 0.1) m.

Here is the plot with the ideal mode of the mode cleaner shown in brown. The ideal mode was plotted with the radius of curvature of 18.4. The blue curve is the fit to the vertical beam radius, the purple curve is the fit to the horizontal beam radius, * denotes a data point from the vertical data, and + denotes a data point from the horizontal data.

Attachment 1: mcfit.png
mcfit.png
  3030   Wed Jun 2 03:24:22 2010 KevinUpdatePSL2W Beam Profile

[Rana, Kiwamu, Kevin]

The Innolight 2W beam profile was measured with the beam scan. A W2-IF-1025-C-1064-45P window was used to reflect a small amount of the main beam. A 5101 VIS mirror was used to direct just the beam reflected from the front surface of the W2 down the table (the beam reflected from the back surface of the W2 hit the optic mount for the mirror). A razor blade beam dump was used to stop the main transmitted beam from the W2. The distance from the laser was measured from the front black face of the laser to the front face of the beam scan (this distance is not the beam path length but was the easiest and most accurate distance to measure). The vertical and horizontal beam widths were measured at 13.5% of the maximum intensity (each measurement was averaged over 100 samples). These widths were divided by 2 to get the vertical and horizontal radii.

The mirror was tilted so that the beam was close to parallel to the table. (The center of the beam fell by approximately 2.1 mm over the 474 mm that the measurement was made in).

The measurement was taken with an injection current of 2.004 A and a laser crystal temperature of 25.04°C.

This data was fit to w = sqrt(w0^2+lambda^2*(x-x0)^2/(pi*w0)^2) with lambda = 1064nm with the following results

For the horizontal beam profile:

reduced chi^2 = 4.0

x0 = (-138 ± 3) mm

w0 = (113.0 ± 0.7) µm

For the vertical beam profile:

reduced chi^2 = 14.9

x0 = (-125 ± 4) mm

w0 = (124.0 ± 1.0) µm

In the following plots, the blue curve is the fit to the vertical beam radius, the purple curve is the fit to the horizontal beam radius, * denotes a data point from the vertical data, and + denotes a data point from the horizontal data.

Attachment 1: profile.png
profile.png
Attachment 2: errors.png
errors.png
Attachment 3: Layout.jpg
Layout.jpg
  3040   Wed Jun 2 22:25:39 2010 KevinUpdatePSLLow Power 2W Beam Profile

Koji is worried about thermal lensing introducing errors to the measurement of the 2W beam profile so I measured the profile at a lower power.

I used the same setup and methods used to measure the profile at 2W (see entry). This measurement was taken with an injection current of 1.202 A and a laser crystal temperature of 25.05° C. This corresponds to approximately 600 mW (see power measurement).

The data was fit to  w = sqrt(w0^2+lambda^2*(x-x0)^2/(pi*w0)^2) with the following results

For the horizontal beam profile:

reduced chi^2 = 2.7

x0 = (-203 ± 3) mm

w0 = (151.3 ± 1.0) µm

For the vertical beam profile:

reduced chi^2 = 6.8

x0 = (-223 ± 6) mm

w0 = (167.5 ± 2.2) µm

In the following plots, the blue curve is the fit to the vertical beam radius, the purple curve is the fit to the horizontal beam radius, * denotes a data point from the vertical data, and + denotes a data point from the horizontal data.

The differences between the beam radii for the low power and high power measurements are

Δw0_horizontal = (38.3 ± 1.2) µm

Δw0_vertical = (43.5 ± 2.4) µm

Thus, the two measurements are not consistent. To determine if the thermal lensing is in the laser itself or due to reflection from the W2 and mirror, we should measure the beam profile again at 2W with a razor blade just before the W2 and a photodiode to measure the intensity of the reflection off of the front surface. If this measurement is consistent with the measurement made with the beam scan, this would suggest that the thermal lensing is in the laser itself and that there are no effects due to reflection from the W2 and mirror. If the measurement is not consistent, we should do the same measurement at low power to compare with the measurement described in this entry.


Attachment 1: profile_low.png
profile_low.png
  3041   Wed Jun 2 22:58:04 2010 KevinUpdatePSL2W Second Reflected Beam Profile

[Koji, Kevin]

The profile of the Innolight 2W was previously measured by measuring the reflected beam from the front surface of a W2 window (see entry). To investigate thermal effects, Rana suggested also measuring the profile of the beam reflected from the back surface of the W2.

I used the same setup and methods as were used in the first measurement. The mirror was moved so that only the beam reflected from the back surface of the W2 was reflected from the mirror. This beam was reflected from both the front of the mirror and the back of the mirror. An extra beam dump was positioned to block the reflection from the back of the mirror.

This measurement was made with 2.004 A injection current and 25.04°C laser crystal temperature.

The data was fit to w = sqrt(w0^2+lambda^2*(x-x0)^2/(pi*w0)^2) with the following results

For the horizontal beam profile:

reduced chi^2 = 5.1

x0 = (-186 ± 6) mm

w0 = (125.8 ± 1.4) µm

For the vertical beam profile:

reduced chi^2 = 14.4

x0 = (-202 ± 11) mm

w0 = (132.5 ± 2.7) µm

In the following plots, the blue curve is the fit to the vertical beam radius, the purple curve is the fit to the horizontal beam radius, * denotes a data point from the vertical data, and + denotes a data point from the horizontal data.

The differences between the beam radii for the beam reflected from the front surface and the beam reflected from the back surface are

Δw0_horizontal = (12.8 ± 1.6) µm

Δw0_vertical = (8.5 ± 2.9) µm

So the two measurements are not consistent. This suggests that the passage through the W2 altered the profile of the beam.

Attachment 1: profile_2nd.png
profile_2nd.png
  3042   Thu Jun 3 00:47:17 2010 KevinUpdatePSL2W Beam Profile of Second Reflected Beam

[Koji, Kevin]

The profile of the Innolight 2W was previously measured by measuring the reflected beam from the front surface of a W2 window (see entry). To investigate thermal effects, Rana suggested also measuring the profile of the beam reflected from the back surface of the W2.

I used the same setup and methods as were used in the first measurement. The mirror was moved so that only the beam reflected from the back surface of the W2 was reflected from the mirror. This beam was reflected from both the front of the mirror and the back of the mirror. An extra beam dump was positioned to block the reflection from the back of the mirror.

This measurement was made with 2.004 A injection current and 25.04°C laser crystal temperature.

The data was fit to w = sqrt(w0^2+lambda^2*(x-x0)^2/(pi*w0)^2) with the following results

For the horizontal beam profile:

reduced chi^2 = 5.1

x0 = (-186 ± 6) mm

w0 = (125.8 ± 1.4) µm


For the vertical beam profile:

reduced chi^2 = 14.4

x0 = (-202 ± 11) mm

w0 = (132.5 ± 2.7) µm


In the following plots, the blue curve is the fit to the vertical beam radius, the purple curve is the fit to the horizontal beam radius, * denotes a data point from the vertical data, and + denotes a data point from the horizontal data.

Attachment 1: profile_2nd.png
profile_2nd.png
  3679   Fri Oct 8 12:29:21 2010 KevinUpdateComputersNew Netgear Switch

I removed some old equipment from the rack outside the control room and stacked them next to the filing cabinets in the control room. I also mounted the new Netgear switch in the rack.

  3747   Wed Oct 20 21:33:11 2010 KevinUpdatePSLQuarter Wave Plate Optimization

[Suresh and Kevin]

We placed the quarter wave plate in front of the 2W laser and moved the half wave plate forward. To make both wave plates fit, we had to rotate one of the clamps for the laser. We optimized the angles of both wave plates so that the power in the reflection from the PBS was minimized and the transmitted power through the faraday isolator was maximized. This was done with 2.1 A injection current and 38°C crystal temperature.

Next, I will make plots of the reflected power as a function of half wave plate angle for a few different quarter wave plate rotations.

  3760   Fri Oct 22 03:37:56 2010 KevinUpdatePSLQuarter Wave Plate Measurements

[Koji and Kevin]

We measured the reflection from the PBS as a function of half wave plate rotation for five different quarter wave plate rotations. Before the measurement we reduced the laser current to 1 A, locked the PMC, and recorded 1.1 V transmitted through the PMC. During the measurements, the beam was blocked after the faraday isolator. After the measurements, we again locked the PMC and recorded 1.2 V transmitted. The current is now 2.1 A and both the PMC and reference cavities are locked.

I will post the details of the measurement tomorrow.

  3768   Sat Oct 23 02:25:49 2010 KevinUpdatePSLQuarter Wave Plate Measurements

Quote:

[Koji and Kevin]

We measured the reflection from the PBS as a function of half wave plate rotation for five different quarter wave plate rotations. Before the measurement we reduced the laser current to 1 A, locked the PMC, and recorded 1.1 V transmitted through the PMC. During the measurements, the beam was blocked after the faraday isolator. After the measurements, we again locked the PMC and recorded 1.2 V transmitted. The current is now 2.1 A and both the PMC and reference cavities are locked.

I will post the details of the measurement tomorrow.

I measured the reflected power from the PBS as a function of half wave plate rotation for five different quarter wave plate rotations.

The optimum angles that minimize the reflected power are 330° for the quarter wave plate and 268° for the half wave plate.

The following data was taken with 2.102 A laser current and 32.25° C crystal temperature.

For each of five quarter wave plate settings around the optimum value, I measured the reflected power from the PBS with an Ophir power meter. I measured the power as a function of half wave plate angle five times for each angle and averaged these values to calculate the mean and uncertainty for each of these angles. The Ophir started to drift when trying to measure relatively large amounts of power. (With approximately 1W reflected from the PBS, the power reading rapidly increased by several hundred mW.) So I could only take data near the minimum reflection accurately.

The data was fit to P = P0 + P1*sin^2(2pi/180*(t-t0)) with the angle t measured in degrees with the following results:

lambda/4 angle (°) t0 (°) P0 (mW) P1 (mW) chi^2/ndf V
318 261.56 ± 0.02 224.9 ± 0.5 2016 ± 5 0.98 0.900 ± 0.001
326 266.07 ± 0.01 178.5 ± 0.4 1998 ± 5 16.00 0.918 ± 0.001
330 268.00 ± 0.01 168.2 ± 0.3 2119 ± 5 1.33 0.926 ± 0.001
334 270.07 ± 0.02 174.5 ± 0.4 2083 ± 5 1.53 0.923 ± 0.001
342 273.49 ± 0.02 226.8 ± 0.5 1966 ± 5 1.41 0.897 ± 0.001

where V is the visibility V = 1- P_max/P_min. These fits are shown in attachment 1. We would like to understand better why we can only reduce the reflected light to ~150 mW. Ideally, we would have V = 1. I will redo these measurements with a different power meter that can measure up to 2 W and take data over a full period of the reflected power.

Attachment 1: fits.png
fits.png
  3802   Thu Oct 28 02:01:51 2010 KevinUpdatePSLFilter for 2W Laser

[Rana and Kevin]

I made a low pass filter for the piezo driver for the 2W laser that is now installed. The filter has a pole at 2.9 Hz. The transfer function is shown in attachment 1.

Attachment 2 shows the outside of the filter with the circuit diagram and attachment 2 shows the inside of the filter.

Attachment 1: tf.PDF
tf.PDF
Attachment 2: outside.jpg
outside.jpg
Attachment 3: inside.jpg
inside.jpg
  3818   Fri Oct 29 04:58:04 2010 KevinUpdatePSLPBS Optimization

[Koji and Kevin]

Since there was still a lot of power being reflected from the PBS before the Faraday rotator, I placed another PBS at the reflection from the first PBS to investigate the problem. If everything was ideal, we would expect the PBS to transmit P polarization and reflect S polarization. Thus, if the laser was entirely in the TEM00 mode, with the quarter and half wave plates we should be able to rotate the polarizations so that all of the power is transmitted through the PBS. In reality, some amount of P is reflected in addition to S reducing the power transmitted. (We are not sure what the PBS is since there are no markings on it but CVI says that their cubes should have less than 5% P reflection).

For the following measurements, the laser crystal temperature was 31.8° C, the current was 2.1 A, the half wave plate was at 267° and the quarter wave plate was at 330°. I first measured the power reflected from the first PBS then added the second PBS to this reflected light and measured the transmitted and reflected powers from this PBS with the following results:

reflection from first PBS 127 mW
reflection from second PBS 48 mW
transmission from second PBS 81 mW

This shows that approximately 81 mW of P polarization was being reflected from the first PBS and that there is approximately 48 mW of S polarization that could not be rotated into P with the two wave plates. Attachment 1 shows the shape of the reflected (S polarization) beam from the second PBS. This shows that the S polarization is not in TEM00 and can not be rotated by the wave plates. The transmitted P polarization is in TEM00.

We then rotated the first PBS (in yaw) to minimize the amount of P being reflected. Repeating the above measurement with the current alignment gives

reflection from first PBS 59 mW
reflection from second PBS 52 mW
transmission from second PBS 8.5 mW

Thus by rotating the cube to minimize the amount of P reflected, ~70 mW more power is transmitted through the cube. This adjustment moved the beam path slightly so Koji realigned the Faraday rotator and EOM. The PMC was then locked and the beam was realigned on the PMC. At 2.1 A, the transmission through the PMC is 6.55 V and the reflection is 178 mV. With the PMC unlocked, the reflection is 312 mV. This gives a visibility of 0.43.

Note by KA:
We realigned the beam toward the PMC at 1.0A at first so that we don't cook any parts. Once we get the TEM00 resonance, the steering mirrors were aligned to maximize the PMC transmission. Then the pumping current was increased to 2.1A.

  3890   Thu Nov 11 02:17:27 2010 KevinUpdateElectronicsREFL11 Photodiode Not Working

[Koji and Kevin]

I was trying to characterize the REFL11 photodiode by shining a flashlight on the photodiode and measuring the DC voltage with an oscilloscope and the RF voltage with a spectrum analyzer. At first, I had the photodiode voltage supplied incorrectly with 15V between the +15 and -15 terminals. After correcting this error, and checking that the power was supplied correctly to the board, no voltage could be seen when light was incident on the photodiode.

We looked at the REFL55 photodiode and could see ~200 mV of DC voltage when shining a light on it but could not see any signal at 55 MHz. If the value of 50 ohm DC transimpedance is correct, this should be enough to see an RF signal. Tomorrow, we will look into fixing the REFL11 photodiode.

  3904   Fri Nov 12 02:51:20 2010 KevinUpdateElectronicsPhotodiode Testing

[Jenne and Kevin]

I started testing the REFL55 photodiode. With a light bulb, I saw ~270 mV of DC voltage from the photodiode but still could not see any RF signal. I connected the RF out from the spectrum analyzer to the test input and verified that the circuit was working.

I then set up the AM laser and looked at the laser light with REFL11 and an 1811 photodiode. I was able to see an RF signal and verified that the resonant frequency is 55 MHz.

The current setup is not very reliable because the laser is not mounted rigidly. Next, I will work on making this mounting more reliable and will continue to work on finding an RF signal with a flashlight.

  3944   Thu Nov 18 01:52:58 2010 KevinUpdateElectronicsREFL55 Transfer Functions

I measured the optical and electrical transfer functions for REFL55 and calculated the RF transimpedance. To measure the optical transfer function, I used the light from an AM laser to simultaneously measure the transfer functions of REFL55 and a New Focus 1611 photodiode. I combined these two transfer functions to get the RF transimpedance for REFL55. I also measured the electrical transfer function by putting the RF signal from the network analyzer in the test input of the photodiode.

I put all of the plots on the wiki at http://lhocds.ligo-wa.caltech.edu:8000/40m/Electronics/REFL55.

  3952   Fri Nov 19 03:43:33 2010 KevinUpdateElectronicsREFL55 Characterizations

[Koji, Rana, and Kevin]

I have been trying to measure the shot noise of REFL55 by shining a light bulb on the photodiode and measuring the noise with a spectrum analyzer. The measured dark noise of REFL55 is 35 nV/rtHz. I have been able to get 4 mA of DC current on the photodiode but have not been able to see any shot noise.

I previously measured the RF transimpedance of REFL55 by simultaneously measuring the transfer functions of REFL55 and a new focus 1611 photodiode with light from an AM laser. By combining these two transfer functions I calculated that the RF transimpedance at 55 MHz is ~ 200 ohms. With this transimpedance the shot noise at 4 mA is only ~ 7 nV/rtHz and would not be detectable above the dark noise.

The value of 200 ohms for the transimpedance seems low but it agrees with Alberto's previous measurements. By modeling the photodiode circuit as an RLC circuit at resonance with the approximate values of REFL55 (a photodiode capacitance of 100 pF and resistance of 10 ohms and an inductance of 40 nH), I calculated that the transimpedance should be ~ 230 ohms at 55 MHz. Doing the same analysis for the values of REFL11 shows that the transimpedance at 11 MHz should be ~ 2100 ohms. A more careful analysis should include the notch filters but this should be approximately correct at resonance and suggests that the 200 ohm measurement is correct for the current REFL55 circuit.

  3971   Tue Nov 23 01:27:33 2010 KevinUpdateElectronicsPOX Characterizations

I measured the RF transimpedance of the POX photodiode by measuring the optical transfer function with the AM laser and by measuring the shot noise with a light bulb. The plots of these measurements are at http://lhocds.ligo-wa.caltech.edu:8000/40m/Electronics/POX.

I measured the noise of the photodiode at 11 MHz for different light intensities using an Agilent 4395a. The noise of a 50 ohm resistor as measured by this spectrum analyzer is 10.6 nV/rtHz. I fit this noise data to the shot noise formula to find the RF transimpedance at 11 MHz to be (2.42 ± 0.08) kΩ. The RF transimpedance at 11 MHz as measured by the transfer function is 6.4 kΩ.

  4048   Mon Dec 13 21:03:30 2010 KevinUpdateElectronicsRF Photodiode Characterizations

[Koji, Jenne, Kevin]

Jenne worked on fixing REFL11 last week (see elog 4034) and was able to measure an electrical transfer function. Today, I tried to measure an optical transfer function but REFL11 is still not responding to any optical input. I tried shining both the laser and a flashlight on the PD but could not get any DC voltage.

I also completed the characterizations of POX. I redid the optical transfer function and shot noise measurements. I also took a time series of the RF output from the PD when it was powered on with no light. This measurement shows oscillations at about 225 MHz. I also measured the spectrum with no light which also shows the oscillations at 225 MHz and smaller oscillations at ~455 MHz.

The plots can be found at http://lhocds.ligo-wa.caltech.edu:8000/40m/Electronics/POX?action=show.

  4167   Wed Jan 19 04:25:54 2011 KevinUpdateElectronicsPOX Transfer Functions

I redid the optical POX transfer functions and updated the wiki at http://lhocds.ligo-wa.caltech.edu:8000/40m/Electronics/POX.

I measured each transfer function several times to calculate uncertainties for each measured point. There is one large transfer function from 1 MHz to 500 MHz showing a resonance peak at 11 MHz and notches at 22 MHz and 55 MHz. I also made more detailed measurements around each of these resonance peaks. These measurements were fit to a resonance curve to determine the resonant frequency, transimpedance at resonance, and Q for each peak. These measurements agree with the shot noise measurement for the transimpedance at 11 MHz taken earlier considering that this measurement was made at 11 MHz instead of at the resonant frequency of 11.14 MHz.

I measured these transfer functions with the Agilent 4395a using the netgpib.py script last week. I realized that when using this script to save multiple copies of the same measurement after setting up the instrument, the first and second measurements are saved but all measurements saved after are identical to the second measurement until the instrument is physically reset. This happens because the analyzer switches the trigger from continuous to hold after making a measurement using this script. Kiwamu said that the script can be modified to return the trigger to continuous after saving the data so that multiple measurements can be saved without being at the analyzer physically. I did not want to waste more time figuring out how to modify the script to do this so I used one of the netbooks and sat at the analyzer manually returning the trigger to continuous after each measurement.

  4170   Wed Jan 19 17:00:23 2011 KevinUpdateElectronicsPOX Transfer Functions

 The value of I_dc was a mistake. The value should be 240 µA.

The widths of the resonance peaks are listed below the fits to each peak on the wiki.

  4172   Thu Jan 20 01:50:30 2011 KevinUpdateElectronicsPOX Transfer Functions

[Koji, Kevin]

We fit the entire POX optical transfer function from 1 MHz to 500 MHz in LISO. The fit is on the wiki at http://lhocds.ligo-wa.caltech.edu:8000/40m/Electronics/POX. Using LISO's root fitting mode, we found that the transfer function has five poles and four zeros.

I will work on making plots of the residuals. This is difficult because by default, LISO does not calculate the fitting function at the frequencies of the data points themselves and I haven't figured out how to force it to do this yet.

ELOG V3.1.3-