40m QIL Cryo_Lab CTN SUS_Lab CAML OMC_Lab CRIME_Lab FEA ENG_Labs OptContFac Mariner WBEEShop
  40m Log, Page 293 of 354  Not logged in ELOG logo
ID Dateup Author Type Category Subject
  14721   Tue Jul 2 19:36:18 2019 aaronUpdateIOOIMC diagnostics

The latest in my fling with the PMC. Though PMC trans is back to nominal levels (~0.713 V), we'd still like to understand the PMC noise.

Last time, I took some spectra with the RF probe (Agilent 41800A). I had already measured the PDH error signal by sweeping the PZT at ~1 Hz. The notebook I used for analysis has been updated in /users/aaron/analysis/PDH_calibrate.ipynb. The analysis was the following:

  • fit the PDH error signal, assuming a 35.5MHz modulation frequency. Here are the (approximate) fit parameters:
    • Mapping of PZT mon voltage to Hz: 5.92 Hz/V_{PZT_mon}
    • P_carrier*P_signal: 0.193 W^2
    • HV mon voltage on resonance: 0.910 V
    • Error signal far off resonance: 0.249 V
    • Transmission: 0.00238
      • ​yikes. The nominal transmission is T=0.003. I let this parameter be free as a check, and to avoid overconstraining the data; is this consistent with measurements of the PMC optics' transmission?
    • Length: 0.0210 m
      • This is consistent with the nominal PMC length
  • Using the fit of the full PDH error signal, I am able to plot error signal vs frequency, and fit the linear portion of the carrier PDH signal. The results of this fit are:
    • -9.75e-7 V_PDH per Hz
    • 0.105 V error signal at DC
  • I then divide the power spectra by the squared slope of the linear fit above (V_PDH^2/Hz^2) to get the spectra in Hz^2/rtHz
    • I've plotted both the spectrum I took directly at the mixer I using the agilent probe, as well as the spectrum taken by sending the PMC servo card's mixer mon to an SR560 (G=2) then to the spectrum analyzer

There are a few problems remaining:

  • There should be a gain of 100 between the mixer I and the servo board's mixer out. It's not clear to me that this is reflected in the spectra. Moreover, the header files on the spectra I grabbed from the Agilent say that the R (mixer I) channel has 20dB of input attenuation, which is also not reflected. If I have swapped the two spectra and not accounted for either the gain of the servo card or the attenuation of the spectrum analyzer, these two gains would cancel, but I'm not confident that's what's going on.
Attachment 1: PDH_error.pdf
PDH_error.pdf
Attachment 2: PMC_Error_Spectrum.pdf
PMC_Error_Spectrum.pdf
  14722   Wed Jul 3 11:47:36 2019 gautamUpdateBHDPRC filtering

A question was raised as to how much passive filtering we benefit from if we pick off the local oscillator beam for BHD from the PRC. I did some simplified modeling of this. For the expected range of arm cavity round trip losses (20-50 ppm), I think that the 40m CARM pole will be between 75-85 Hz. The corresponding recycling gain will be 40-50, with the current PRM. I assumed 1000 ppm loss inside the PRC. The net result is that, assuming the single pole coupled cavity response, we will get ~8-9 dB of filtering at ~200 Hz of the intensity noise of the input laser field to the interferometer if we pick the LO beam off from the PRC (e.g. PR2 transmission), instead of picking it off before.

The next questions are: (i) can we do a sufficiently good job of achieving the required RIN stability on the LO field for BHD without relying on the passive filtering action of the PRC? and (ii) is the benefit of the PRC filtering ruined in the process of routing the LO field from wherever the pickoff happens to the BHD setup?

Attachment 1: PRCfiltering.pdf
PRCfiltering.pdf
  14723   Wed Jul 3 23:53:38 2019 MilindUpdateCamerasdata for nns

Tried collecting data today. Was unable to keep the camera_server code running for any length of time as it threw segfaults. Will take a shot again tomorrow.

Quote:

The GigE is focused now (judged by eye) and I have closed the lid. I'm attaching a picture of the MC2 beam spot, captured using GigE at an exposure time of 400µs.

What was the solution to resolving the flaky video streaming during the alignment process????

-> I think, the issue was with either the poor wireless network conection or the GigE-PoE ethernet cable.

Quote:

Turns out, focusing the GigE is actually a bit tricky. With pylon, everytime I change the exposure or the focus, I'm running into the error I had mentioned earlier in one of my elogs; so I tried using the python scripts to interact with the GigE. But whenever I try to change the focal plane distance by rotating the lens coupler, the ethernet cable connection becomes loose and the camera server needs to be relaunched every now and then. Also, everytime we want to change the distance between the lenses, the telescope needs to be dismantled and refocused again. I'll try to come up with a better telescope design for this.

Yesterday, I had focused the GigE using a low exposure time and small aperture of iris, to make sure that we are actually seeing a sharp image of the beam spot. I'm attaching a picture of the beam spot I had clicked while focusing it, unfortunately, I forgot to take a picture after I had focused it completely. I'm also attaching a picture of the final setup for future reference. 


Yesterday night, Rana asked me to lock the MC2. I figured that the PSL shutter was closed; I just opened it and was able to see the beam spot on the analog camera screen.

  14724   Thu Jul 4 10:47:37 2019 MilindUpdateGeneralEarthquake now

There was a magnitude 6.6 earthquake just a few minutes ago. I am attaching photographs of the monitor feeds for reference here. Is there a standard protocol to be followed in this situation? I'm looking through the wiki now.

Further, the IMC seems to be misaligned and is not locking! cryingcrying As Koji has let me know, I really hope this is not too serious and can be fixed easily.

Attachment 1: after_earthquake2.jpg
after_earthquake2.jpg
Attachment 2: after_earthquake.jpg
after_earthquake.jpg
  14725   Thu Jul 4 10:54:21 2019 KojiSummarySUSSuspension damping recovered, ITMX stuck

So Cal Earthquake. All suspension watchdogs tripped.

Tried to recover the OSEM damping. 

=> The watchdogs for all suspensions except for ITMX were restored. ITMX seems to be stuck. No further action by me for now.

  14726   Thu Jul 4 18:19:08 2019 MilindUpdateCamerasConvolutional neural networks for beam tracking

The quoted elog has figures which indicate that the network did not learn (train or generalize) on the used data. This is a scary thing as (in my experience) it indicates that something is fundamentally wrong with either the data or model and learning will not happen despite how hyperparameters are tuned. To check this, I ran the training experiment for nearly 25 hyperparameter settings (results here)with the old data and was able to successfully overfit the data. Why is this progress? Well, we know that we are on the right track and the task is to reduce overfitting. Whether, that will happen through more hyperparameter tuning, data collection or augmentation remains to be seen. See attachments for more details. 

Why is the fit so perfect at the start and bad later? Well, that's because the first 90% of the test data is  the training data I overfit to and the latter the validation data that the network has not generalized well to.

Quote:

And finally, a network is trained!

Result summary (TLDR :-P) : No memory was used. Model trained. Results were garbage. Will tune hyperparameters now. Code pushed to github.

 

More details of the experiment:

Aim:

  1. To train a network to check that training occurs and get a feel for what the learning might be like.
  2. To set up the necessary framework to perform mulitple experiments and record results in a manner facilitating comparison.
  3. To track beam spot motion.

What I did:

  1. Set up a network that learns a framewise mapping as described in here.
  2. Training data: 0.9 x 1791 frames. Validation data: 0.1 x 1791 frames. Test data (only prediction): all the 1791 frames
  3. Hyperparameters: Attachment #1
  4. Did no tuning of hyperparameters.
  5. Compiled and fit the models and saved the results.

 

What I saw

  1. Attachment #2: data fed to the network after pre-processing - median blur + crop
  2. Attachment #3: learning curves.
  3. Attachment #4: true and predicted motion. Nothing great.

What I think is going wrong-

  1. No hyperparameter tuning. This was only a first pass but is being reported as it will form the basis of all future experiments.
  2. Too little data.
  3. Maybe wrong architecture.

Well, what now?

  1. Tune hyperparmeters (try to get the network to overfit on the data and then test on that. We'll then know for sure that all we probably need is more data?)
  2. Currently the network has around 200k parameters. Maybe reduce that.
  3. Set up a network that takes as input (one example corresponding to one forward pass)  a bunch of frames and predicts a vector of position values that can be used as continuous data).
Quote:

I got to speak to Gabriele about the project today and he suggested that if I am using Rana's memory based approach, then I had better be careful to ensure that the network does not falsely learn to predict a sinusoid at all points in time and that if I use the frame wise approach I try to somehow incorporate the fact that certain magnitudes and frequencies of motion are simply not physically possible. Something that Rana and Gautam emphasized as well.

Quote:
 
  1. Network training for beam spot tracking: I will begin training the convolutional network with the data pre-processed as described above. I will also simultaneously prepare data acquired from the GigE and train networks on that. Note: I planned to experiment with framewize predictions and hence did some of the work described above. However, I will restrict the number of experiments on that and perform more of those that use 3D convolution. Rana also pointed out that it would be interesting to have the network output uncertainity in the predictions. I am not sure how this can be done, but I will look into it.
Attachment 1: Motion.pdf
Motion.pdf
Attachment 2: Error.pdf
Error.pdf
Attachment 3: Learning_curves.pdf
Learning_curves.pdf
  14727   Fri Jul 5 20:57:04 2019 KojiUpdateSUSAnother M7.1 EQ

[Kruthi, Koji]

Koji came to the lab to align the IMC/IFO, but found the mirrors are dancing around. Kruthi told me that there was M7.1 EQ at Ridgecrest. Looks like there are aftershocks of this EQ going on. So we need to wait for an hour to start the alignment work.

ITMX and ETMX are stuck.

Attachment 1: Screenshot_from_2019-07-05_21-03-06.png
Screenshot_from_2019-07-05_21-03-06.png
  14728   Fri Jul 5 21:53:10 2019 KojiUpdateSUSAnother M7.1 EQ

- ITM unstuck now
- IMC briefly locked at TEM00

A series of aftershocks came. I could unstick ITMX by turning on the damping during one of the aftershocks.
Between the aftershocks, MC1~3 were aligned to the previous dof values. This allowed the IMC flashing. Once I got the lock of a low order TEM mode, it was easy to recover the alignment to have a weak TEM00.
Now at least temporarily the full alignment of the IMC was recovered.

  14729   Fri Jul 5 22:21:13 2019 KojiUpdateSUSAnother M7.1 EQ

In fact, ETMX was not stuck until the M7.1 EQ today. After that it got stuck, but during the after shocks, all the OSEMs occasionally showed full swing of the light levels. So I believe the magnets are OK.

Attachment 1: Screenshot_from_2019-07-05_22-19-57.png
Screenshot_from_2019-07-05_22-19-57.png
  14730   Fri Jul 5 23:28:52 2019 rana, kruthiSummarySUSETMX unstuck by shaking the stack

We unstuck ETMX by shaking the stack. Most effective was to apply large periodic human sized force to the north STACIS mounts.

At first, we noticed that the face OSEMs showed nearly zero variation.

We tried unsticking it through the usual ways of putting large excitations through AWG into the pit/yaw/side DOFs. This produced only ~0.2 microns of motion as seen by the OSEMs.

After the stack shake, we used the IFO ALIGN sliders to get the oplev beam back on the QPD.

The ETMX sensor trends observed before and after the earthquake are attached.

** plots deleted; SOMEONE, tried to take raster images and turn them into PDF as if this would somehow satisfy our vetor graphics requirement. Boo. lpots must be actual vector graphics PDF

  14731   Sun Jul 7 17:54:34 2019 MilindUpdateComputer Scripts / ProgramsPMC autolocker

I modified the autolocker code I wrote to read from a .yaml configuration file instead of commandline arguements (that option still exists if one wishes to override what the .yaml file contains). I have pushed the code to github. I started reading about MCMC and will put up details of the remaining part of the work ASAP.

Quote:
 

P.P.S.  He also said that it would not do to have command line arguments as the main source from which parameters are procured and that .yml files ought to be used instead. I will make that change asap.

  14732   Sun Jul 7 21:59:28 2019 KruthiUpdateCamerasGhost image due to beamsplitter

The beam splitter (BS1-1064-33-2037-45S) that is currently being used has an antireflection coating on the second surface and a wedge of less than 5 arcmin; yet it leads to ghosting as shown in the figure attached (courtesy: Thorlabs). I'm also attaching its spec sheet I dug up on internet for future reference.

I came across pellicle beamsplitters, that are primarily used to eliminate ghost images. Pellicle beamsplitters have a few microns thick nitrocellulose layer and superimpose the secondary reflection on the first one. Thus the ghost image is eliminated. 

Should we go ahead and order them? (https://www.thorlabs.com/newgrouppage9.cfm?objectgroup_id=898

https://www.edmundoptics.eu/c/beamsplitters/622/#28438=28438_s%3AUGVsbGljbGU1&27614=27614_d%3A%5B46.18%20TO%2077.73%5D)

Attachment 1: ghosting_schematic.png
ghosting_schematic.png
Attachment 2: Beamsplitter_spec.pdf
Beamsplitter_spec.pdf Beamsplitter_spec.pdf
  14733   Mon Jul 8 17:33:10 2019 KruthiUpdateLoss MeasurementOptical scattering measurements

I came across a paper (see reference) where they have used DAOPHOT, an astronomical software tool developed by NOAO, to study the point scatterers in LIGO test masses using images of varying exposure times. I'm going through the paper now. I think using this we can analyze the MC2 images and make some interesting observations.

Reference:  L.Glover et al., Optical scattering measurements and implications on thermal noise in Gravitational Wave detectors test-mass coatings Physics Letters A. 382. (2018)

  14734   Mon Jul 8 17:52:30 2019 MilindUpdateCamerasConvolutional neural networks for beam tracking

After the two earthquakes, I collected some data by dithering the optic and recording the QPD readings. Today, I set up scripts to process the data and then train networks on this data. I have pushed all the code to github. I attempted to train a bunch of networks on the new data to test if the code was alright but realised quickly that, training on my local machine is not feasilble at all as training for 10 epochs took roughly 6 minutes. Therefore, I have placed a request for access to the cluster and am waiting for a reply. I will now set up a bunch of experiments to tune hyperparameters for this data and see what the results are.

Trainng networks with memory

I set up a network to handle input volumes (stacks of frames) instead of individual frames. It still uses 2D convolution and not 3D convolution. I am currently training on the new data. However, I was curious to see if it would provide any improved performance over the results I put up in the previous elog. After a bit of hyperparameter tuning, I did get some decent results which I have attached below. However, this is for Pooja's old data which makes them, ah, not so relevant. Also, this testing isn't truly representative because the test data isn't entirely new to the network. I am going to train this network on the new data now with the following objectives (in the following steps):

  1. Train on data recorded at one frequency, generalize/ test on unseen data of the same frequency, large amplitude of motion
  2. Train on data recorded at one frequency, generallize/ test on unseen data of a different frequency, large amplitude of motion
  3. Train on data recorded at one frequency, generalize/ test on unseen data of  same/ different frequency, small amplitude of motion
  4. Train on data at different frequencies and generalize/ test on data with a mixture of frequencies at small amplitudes - Gautam pointed out that the network would truly be superb (good?) if we can just predict the QPD output from the video of the beam spot when nothing is being shaken.

I hope this looks alright? Rana also suggested I try LSTMs today. I'll maybe code it up tomorrow. What I have in mind- A conv layer encoder, flatten, followed by an LSTM layer (why not plain RNNs? well LSTMs handle vanishing gradients, so why the hassle).

Quote:

The quoted elog has figures which indicate that the network did not learn (train or generalize) on the used data. This is a scary thing as (in my experience) it indicates that something is fundamentally wrong with either the data or model and learning will not happen despite how hyperparameters are tuned. To check this, I ran the training experiment for nearly 25 hyperparameter settings (results here)with the old data and was able to successfully overfit the data. Why is this progress? Well, we know that we are on the right track and the task is to reduce overfitting. Whether, that will happen through more hyperparameter tuning, data collection or augmentation remains to be seen. See attachments for more details. 

Why is the fit so perfect at the start and bad later? Well, that's because the first 90% of the test data is  the training data I overfit to and the latter the validation data that the network has not generalized well to.

Attachment 1: Motion.pdf
Motion.pdf
  14735   Mon Jul 8 21:42:39 2019 ranaUpdateCamerasGhost image due to beamsplitter

you have to use a BS with a larger wedge angle (5 arcmin ~ 1 mrad) so that the beams don't overlap on the camera

  14736   Tue Jul 9 08:33:31 2019 gautamSummarySUSETMX PIT bias voltage changed by ~1V

After this activity, the DC bias voltage required on ETMX to restore good X arm cavity alignment has changed by ~1.3 V. Assuming a full actuation range of 30 mrad for +/- 10 V, this implies that the pitch alignment of the stack has changed by ~2 mrad? Or maybe the suspension wires shifted in the standoff grooves by a small amount? This is ~x10 larger than the typical change imparted while working on the table, e.g. during a vent.

Main point is that this kind of range requirement should probably be factored in when thinking about the high-voltage coil driver actuation.

Quote:

We unstuck ETMX by shaking the stack. Most effective was to apply large periodic human sized force to the north STACIS mounts.

  14737   Tue Jul 9 10:37:42 2019 MilindUpdateIOOkeyed psl crate, unstick.py, pmc autolocker code- working

Today, Gautam keyed the C1PSL crate and we got to test my unstick.py code. It seems to be working fine. Remarks:

  1. Gautam moved the unstick.py code to /opt/rtcds/caltech/c1/scripts/cds. Therefore, the steps to run this code are now:
    1. cd /opt/rtcds/caltech/c1/scripts/cds
    2. python unstick.py c1psl (for the c1psl machine)
  2. There is now a sleepTime global variable in the code which defines the amount of delay between successive channel toggles. We set this to 1ms and it took the code around 3s to run.
  3. Gautam was curious to see if this would work even if we set the sleepTime parameter to 0 but decided that that could be tested the next time something was keyed.
  4. I still need to add the signal handling thing to this code.

Following this, we tested my PMC autolocker code. The code ran for about a minute before achieveing lock. Remarks:

  1. Gautam moved my code (pmc_autolocker.py and autolocker_config.yaml) to /cvs/cds/rtcds/caltech/c1/scripts/PSL/PMC/ . Therefore, the steps to run this code are now:
    1. cd /cvs/cds/rtcds/caltech/c1/scripts/PSL/PMC/
    2. python pmc_autolocker.py (check code or use --help to see what the command line arguments do which is only for when you wanna override the details in the .yaml file)
  2. Gautam suggested that I add some delay between succesive steps of DC output adjust so that it locks quickly. I'll do that ASAP. For now, it works.
  14738   Tue Jul 9 18:06:05 2019 gautamUpdateLSCY-arm ASS in a workable state

The Y-arm ASS was tuned to be in a workable state. Basically, I followed Koji's recipe.

The SNR of the dither lines in the TRY and YARM control signals were checked - Attachment #1. The dither frequencies are marked with vertical dashed lines (can't figure out how to add 4 cursors in DTT so there's two in each row for a total of 4). A couple of days ago, when I was doing some preliminary checks, I found that the oscillator at 24.91 Hz caused a broadband increase in the TRY noise between DC and ~100 Hz. But today I saw no evidence of such behaviour. So I decided against changing the frequency.

The linearity of the demodulated error signals around the quadratic maxima of the TRY level was checked. I did not, however, investigate in detail the frequency-dependent offset Koji has reported in his elog. 

After this work, the TRY level is at 0.95. This is commensurate with the MC trans level being lower by ~7% relative to July 2018. Furthermore, the ASS servo is able to return to TRY~0.95 with a time-constant of ~5 seconds in response to misalignment of the cavity optics. After I investigate the X-arm ASS, I will reset the normalization for TRX and TRY.

Update 645pm: In the spirit of general IFO recovery, I re-centered the ITM and ETM oplev spots, and also the IR beam on the IPPOS QPD to mark the new input pointing alignment (the spot is slightly lower on the AS camera than what I remember). I then tweaked the XARM transmission to maximize it, and re-set the TransMon normalization. I edited the normalization script to comment out the normalizing of the TransMon QPD gains as the QPDs are in some kind of indeterminate state now. Attachment #2 shows the current status, you can also see the normalization being reset. LSC mode disabled for overnight.

Once the XARM ASS is also checked out, I propose moving back to locking the DRMI / PRFPMI configs. 

Attachment 1: ditherFreqs.pdf
ditherFreqs.pdf
Attachment 2: transRenorm.png
transRenorm.png
  14739   Tue Jul 9 18:17:48 2019 gautamUpdateGeneralProjector lightbulb blown out

Last documented replacement in Nov 2018, so ~7 months, which I believe is par for the course. I am disconnecting its power supply cable.

  14740   Tue Jul 9 18:42:15 2019 gautamUpdateALSEX green doubling oven temperature controller power was disconnected

There was no green light even though the EX NPRO was on. I checked the doubling oven temperature controller and found that its power cable was loose on the rear. I reconnected it, and now there is green light again. 

  14741   Tue Jul 9 22:13:26 2019 MilindUpdateCamerasConvolutional neural networks for beam tracking

I received access today. After some incredible hassle, I was able to set up my repository and code on the remote system. Following this, Gautam wrote to Gabriele to ask him about which GPUs to use and if there was a previously set up environment I could directly use. Gabriele suggested that I use pcdev2 / pcdev3 / pcdev11 as they have good gpus. He also said that I could use source ~gabriele.vajente/virtualenv/bin/activate to use a virtualenv with tensorflow, numpy etc. preinstalled. However, I could not get that working, Therefore I created my own virtual environment with the necessary tensorflow, keras, scipy, numpy etc. libraries and suitable versions. On ssh-ing into the cluster, it can be activated using source /home/millind.vaddiraju/beamtrack/bin/activate. How do I know everything works? Well, I trained a network on it! With the new data. Attached (see attachment #1) is the prediction data for completely new test data. Yeah, its not great, but I got to observe the time it takes for the network to train for 50 epochs-

  1. On pcdev5 CPU: one epoch took ~1500s which is roughly 25 minutes (see Attachment #2). Gautam suggested that I try to train my networks on Optimus. I think this evidence should be sufficient to decide against that idea.
  2. On my GTX 1060: one epoch took ~30s. Which is 25 minutes (for 50 epochs) to train a network.
  3. On pcdev11 GPU (Titan X I think): each epoch took ~16s which is a far more reasonable time.

Therefore, I will carry out all training only on this machine from now.

 


Note to self:

Steps to repeat what you did are:

  1. ssh in to the cluster using ssh albert.einstein@ssh.ligo.org as described here.
  2. activate virtualenv as descirbed above
  3. navigate to code  and run it.
Quote:

 I attempted to train a bunch of networks on the new data to test if the code was alright but realised quickly that, training on my local machine is not feasilble at all as training for 10 epochs took roughly 6 minutes. Therefore, I have placed a request for access to the cluster and am waiting for a reply. I will now set up a bunch of experiments to tune hyperparameters for this data and see what the results are.

Attachment 1: predicted_motion_first.pdf
predicted_motion_first.pdf
Attachment 2: pcdev5_time.png
pcdev5_time.png
  14742   Wed Jul 10 10:04:09 2019 gautamUpdateSUSTip-Tilt moved from South clean cabinet to bake lab cleanroom

Arnaud and I moved one of the two spare TT suspensions from the south clean cabinet to the bake lab clean room. The main purpose was to inspect the contents of the packaging. According to the label, this suspension was cleaned to Class A standards, so we tried to be clean while handling it (frocks, gloves, masks etc). We found that the foil wrapping contained one suspension cage, with what looked like all the parts in a semi-assembled state. There were no OSEMs or electronics together with the suspension cage. Pictures were taken and uploaded to gPhoto. Arnaud is going to plan his tests, so in the meantime, this unit has been stored in Cabinet #6 in the bake lab cleanroom.

  14743   Wed Jul 10 14:55:32 2019 KojiUpdateGeneralProjector lightbulb blown out

In fact the projector is still working. The lamp timer showed ~8200hrs. I just reset the timer, but not sure it was the cause of the shutdown. I also set the fan mode to be "High Altitude" to help cooling.

  14744   Wed Jul 10 14:57:01 2019 KojiSummaryCDSChannel recipe for iscaux upgrade

The list of the iscaux channels and pin assignments were posted to google drive.
The spreadsheet can be viewable by the link sent to the 40m ML. It was shared with foteee@gmail for full access.

Summary

  • We need
    4 ADC modules
    5 DAC modules
    5 Binary I/O modules
  • Be aware that there are bundled multiple digital I/O channels such as "mbboDirect" and "mbbi".
  • The full db record of the new channels need to be inferred from the existing channels.

Necessary electronics modification

1. D990694 whitening filter modification (4 modules)

This module shares the fast and slow channels on the top DIN96pin (P1) connector. Also, the whitening selector (done by an analog signal per channel) is assigned over 17pin of the P1 connector, resulting in the necessity of the second DSUB cable. By migrating the fast channels, we can swap the cable from the P1 to P2.  Also, the whitening selectors are concentrated on the first Dsub. (See Attachment1 P1)

2. D040180 / D1500308 Common Mode Board

CM servo board itself doesn't need any modification. The CM board uses P1 and P2. So we need to manufacture a special connector for CM Board P2. (cf The adapter board for P1 T1800260). See also D1700058.

3. D990543A1 LSC Photodiode Interface

PD I/F board has the DC mon channels spread over the 16pin limit. P1 21A can be connected to 6A so that we can accommdate it in the first Dsub.
Also the board uses AD797s. This is not necessary. We can replace them to OP27s. I actually don't know what is happening to those bias control, temp mon, enable, and status. These features should be disables at the I/F and the PDs. (See Attachment2 P1)

Attachment 1: D990694-B.pdf
D990694-B.pdf D990694-B.pdf D990694-B.pdf D990694-B.pdf D990694-B.pdf D990694-B.pdf D990694-B.pdf D990694-B.pdf
Attachment 2: D990543A1.PDF
D990543A1.PDF D990543A1.PDF D990543A1.PDF
  14745   Wed Jul 10 16:53:22 2019 gautamUpdateSUSPRM watchdog condition modified

[koji, gautam]

We noticed that the PRM watchdog was tripping frequently. This is a period of enhanced seismic activity. The reason PRM in particular trips often is because the SIDE OSEM has 5x increased transimpedance. We implemented a workaround by modifying the watchdog tripping condition to scale the SD channel RMS by a factor of 0.2 (relative to the UL and LL channels). We restarted the modbus process on c1susaux and tested that the new logic works. Here is the relevant snippet of code:

# Disable fast DAC if variation tests too high
# PRM Side is special, see elog 14745
record(calc,"C1:SUS-PRM_LOGIC")
{
    field(DESC,"Tests whether RMS too high")
    field(SCAN,"1 second")
    field(PHAS,"1")
    field(PREC,"0")
    field(HOPR,"1")
        field(LOPR,"0")
        field(CALC,"(A<B)&(C<B)&(0.2*D<B)")
        field(INPA,"C1:SUS-PRM_ULPD_VAR  NPP  NMS")
        field(INPB,"C1:SUS-PRM_PD_MAX_VAR  NPP  NMS")
        field(INPC,"C1:SUS-PRM_LLPD_VAR  NPP  NMS")
        field(INPD,"C1:SUS-PRM_SDPD_VAR  NPP  NMS")
}

The db file has a note about this as well so that future debuggers aren't mystified by a factor of 0.2.

  14746   Wed Jul 10 22:32:38 2019 MilindUpdateCamerasConvolutional neural networks for beam tracking

I trained a bunch (around 25 or so - to tune hyperparameters) of networks today. They were all CNNs. They all produced garbage. I also looked at lstm networks with CNN encoders (see this very useful link) and gave some thought to what kind of architecture we want to use and how to go about programming it (in Keras, will use tensorflow if I feel like I need more control). I will code it up tomorrow after some thought and discussion. I am not sure if abandoning CNNs is the right thing to do or if I should continue probing this with more architectures and tuning attempts. Any thoughts?

Right now, after speaking to Stuart (ldas_admin) I've decided on coding up the LSTM thing and then running that on one machine while probing the CNN thing on another.

 


Update on 10 July, 2019: I'm attaching all the results of training here in case anyone is interested in the future.

Quote:

I received access today. After some incredible hassle, I was able to set up my repository and code on the remote system. Following this, Gautam wrote to Gabriele to ask him about which GPUs to use and if there was a previously set up environment I could directly use. Gabriele suggested that I use pcdev2 / pcdev3 / pcdev11 as they have good gpus. He also said that I could use source ~gabriele.vajente/virtualenv/bin/activate to use a virtualenv with tensorflow, numpy etc. preinstalled. However, I could not get that working, Therefore I created my own virtual environment with the necessary tensorflow, keras, scipy, numpy etc. libraries and suitable versions. On ssh-ing into the cluster, it can be activated using source /home/millind.vaddiraju/beamtrack/bin/activate. How do I know everything works? Well, I trained a network on it! With the new data. Attached (see attachment #1) is the prediction data for completely new test data. Yeah, its not great, but I got to observe the time it takes for the network to train for 50 epochs-

  1. On pcdev5 CPU: one epoch took ~1500s which is roughly 25 minutes (see Attachment #2). Gautam suggested that I try to train my networks on Optimus. I think this evidence should be sufficient to decide against that idea.
  2. On my GTX 1060: one epoch took ~30s. Which is 25 minutes (for 50 epochs) to train a network.
  3. On pcdev11 GPU (Titan X I think): each epoch took ~16s which is a far more reasonable time.

Therefore, I will carry out all training only on this machine from now.

 


Note to self:

Steps to repeat what you did are:

  1. ssh in to the cluster using ssh albert.einstein@ssh.ligo.org as described here.
  2. activate virtualenv as descirbed above
  3. navigate to code  and run it.
Quote:

 I attempted to train a bunch of networks on the new data to test if the code was alright but realised quickly that, training on my local machine is not feasilble at all as training for 10 epochs took roughly 6 minutes. Therefore, I have placed a request for access to the cluster and am waiting for a reply. I will now set up a bunch of experiments to tune hyperparameters for this data and see what the results are.

  14747   Thu Jul 11 12:42:35 2019 gautamSummaryCDSP2 interface board

I looked into the design of the P2 interface board. The main difficulty here is geometric - we have to somehow accommodate sufficient number of D-sub connectors in the tight space between the two P-type connectors. 

I think the least painful option is to stick with Johannes' design for the P1 connector. For the CM board, the P2 connector only uses 6 pairs of conductors for signals. So we can use a D-15 connector instead of 2 D-37 connectors. Then we can change the PCB shape such that the P1 connector can be accommodated (see Attachment #1). The other alternative would be to have 2 P-type connectors and 3 D-subs on the same PCB, but then we have to be extra careful about the relative positioning of the P-type connectors (otherwise they wont fit onto the Eurocrate). So I opted to still have two separate PCBs.

I took a first pass at the design, the files may be found here. I just auto-routed the connections, this is just an electrical feedthrough so I don't think we need to be too concerned about the PCB trace routing? If this looks okay, we should send out the piece for fab ASAP.

I will work on putting together the EPICS server machine (SuperMicro) this afternoon.

Quote:

2. D040180 / D1500308 Common Mode Board

CM servo board itself doesn't need any modification. The CM board uses P1 and P2. So we need to manufacture a special connector for CM Board P2. (cf The adapter board for P1 T1800260). See also D1700058.

Attachment 1: IMG_7728.JPG
IMG_7728.JPG
  14749   Thu Jul 11 13:08:36 2019 ChubSummaryCDSP2 interface board

It's nice and compact, and the cost of new 15-pin DSUB cables shouldn't be a factor here.  What does the 15p cable connect to?

  14750   Thu Jul 11 13:09:22 2019 gautam SummaryCDSP2 interface board

it will connect to a 15 pin breakout board in the Acromag chassis

Quote:

It's nice and compact, and the cost of new 15-pin DSUB cables shouldn't be a factor here.  What does the 15p cable connect to?

  14752   Thu Jul 11 16:22:54 2019 KruthiUpdateGeneralProjector lightbulb blown out

I heard a popping sound in the control room; the projector lightbulb has blown out.sad

  14753   Thu Jul 11 17:58:38 2019 gautamUpdateEquipment loanTT suspension --> Downs

Arnaud has taken 1 TT suspension from the 40m clean lab to Downs for modal testing. Estimated time of return is tomorrow evening.

  14754   Thu Jul 11 18:15:22 2019 gautamSummaryElectronicsPSL/IOO rack checkout

I looked at the PSL/IOO racks to check for which boards, if any, require an additional P2 interface, so that we can try and design a generic one for the IMC/CM boards and whatever else may require it. While searching the elog, I saw that Koji and Johannes had already done this, see Koji's elog in this thread. Some remarks:

  1. D990155 seems to be unused in both PSL and IOO racks. The one in the PSL rack has some LEMO cables plugged in to the front panel, but they go nowhere. So I think that both of these are redundant (in the assessment below, only one was marked redundant).
  2. In the PSL rack, the "TTFSS Interface", "PSL PMC SERVO", and "DAQ INTERFACE" (which I think is obsolete) cards all have their P2 connectors daisy chained together, going to a cross-connect. Kruthi and I traced this to be going to a cross connect marked "J23-PSLRACK-CCP". In the PSL wiring diagram of which we have a hardcopy in the control room, it looks like these channels are related to the RefCav? So I think this is not required to be interfaced to our new Acromag DAQ system. 

Conclusion: Only the IMC Servo and CM boards need their P2 connectors connected to Acromag.It would be helpful to remove the TTFSS Interface board and figure out what exactly the pin-mapping for the backplane connectors are, but I didn't do this today because there is a "High Voltage" line going to the Interface Board and I'm not actually sure of the signal chain for the FSS servo.

  14755   Fri Jul 12 07:37:48 2019 gautamUpdateSUSM4.9 EQ in Ridgecrest

All suspension watchdogs were tripped ~90mins ago. I restored the damping. IMC is locked.

ITMX was stuck. I set it free. But notice that the UL Sensor RMS is higher than the other 4? I thought ITMY UL was problematic, but maybe ITMX has also failed, or maybe it's coincidence? Something for IFOtest to figure out I guess. I don't think there is a cable switch between ITMX/ITMY as when I move the ITMX actuators, the ITMX sensors respond and I can also see the optic moving on the camera.

Took me a while to figure out what's going on because we don't have the seis BLRMS - i moved the usual projector striptool traces to the TV screen for better diagnostic ability.

Update 16 July 1515: Even though the RMS is computed from the slow readback channels, for diagnosis, I looked at the spectra of the fast PD monitoring channels (i.e. *_SENSOR_*) for ITMX - looks like the increased UL RMS is coming from enhanced BR-mode coupling and not of any issues with the whitening switching (which seems to work as advertised, see Attachment #3, where the LL traces are meant to be representative of LL, LR, SD and UR channels).

Attachment 1: 56.png
56.png
Attachment 2: ITMXunstick.png
ITMXunstick.png
Attachment 3: ITMX_UL.pdf
ITMX_UL.pdf
  14756   Fri Jul 12 18:54:47 2019 KojiUpdateGeneralItem loan: optical chopper from Cryo Lab

Optical chopper borrowed from CryoLab to 40m

https://nodus.ligo.caltech.edu:8081/Cryo_Lab/2458

  14757   Sun Jul 14 00:24:29 2019 KruthiUpdateCamerasCCD Calibration

On Friday, I took images for different power outputs of LED. I calculated the calibration factor as explained in my previous elog (plots attached).

Vcc (V) Photodiode
reading(V)

Power incident on photodiode (W)

Power incident on GigE (W)
Slope (counts/​𝝁s)
Uncertainity in
 slope (counts/​𝝁s)
CF (W-sec/counts)
16 0.784 2.31E-06 3.89E-07 180.4029 1.02882 2.16E-15
18 0.854 2.51E-06 4.24E-07 207.7314 0.7656 2.04E-15
20 0.92 2.71E-06 4.57E-07 209.8902 1.358 2.18E-15
22 0.969 2.85E-06 4.81E-07 222.3862 1.456 2.16E-15
25 1.026 3.02E-06 5.09E-07 235.2349 1.53118 2.17E-15
  Average  2.14E-15

To estimate the uncertainity, I assumed an error of at most 20mV (due to stray lights or difference in orientation of GigE and photodiode) for the photodiode reading. Using the uncertainity in slope from the linear fit, I expect an uncertainity of maximum 4%. Note: I haven't accounted for the error in the responsivity value of the photodiode.

GigE area 10.36 sq.mm
PDA area 61.364 sq.mm
Responsivity 0.34 A/W
Transimpedance gain (at gain = 20dB) 10^6 V/W +/- 0.1%
Pixel format used Mono 8 bit

Johannes had reported CF as 0.0858E-15 W-sec/counts for 12 bit images, with measured a laser source. This value and the one I got are off by a factor of 25. Difference in the pixel formats and effect of coherence of the light used might be the possible reasons.

Attachment 1: CCD_calibration.png
CCD_calibration.png
  14758   Mon Jul 15 03:15:24 2019 KruthiUpdateLoss MeasurementImaging scatterometer

On Friday, Koji helped me find various components required for the scatterometer setup. Like he suggested, I'll first set it up on the SP table and try it out with an usual mirror. Later on, once I know it's working, I'll move the setup to the flow bench near the south arm and measure the BRDF of a spare end test mass.

  14759   Mon Jul 15 03:30:47 2019 KruthiUpdateCalibration-RepairWhite paper as a Lambertian scatterer

I made some rough measurements, using the setup I had used for CCD calibration, to get an idea of how good of a Lambertian scatterer the white paper is. Following are the values I got:

Angle (degrees) Photodiode reading (V)  Ps (W) BRDF (per str) % error
12 0.864 2.54E-06 0.334 20.5
24 0.926 2.72E-06 0.439 19.0
30 1.581 4.65E-06 0.528 19.0
41 0.94 2.76E-06 0.473 19.8
49 0.545 1.60E-06 0.423 22.5
63 0.371 1.09E-06 0.475 28

Note: All the measurements are just rough ones and are prone to larger errors than estimated.

I also measured the transmittance of the white paper sample being used (it consists of 2 white papers wrapped together). It was around 0.002

Attachment 1: BRDF_paper.png
BRDF_paper.png
  14760   Mon Jul 15 14:09:07 2019 MilindUpdateCamerasCNN LSTM for beam tracking

I've set up network with a CNN encoder (front end) feeding into a single LSTM cell followed by the output layer (see attachment #1). The network requires significantly more memory than the previous ones. It takes around 30s for one epoch of training. Attached are the predicted yaw motion and the fft of the same. The FFT looks rather curious. I still haven't done any tuning and these are only the preliminary results.

Quote:

 Rana also suggested I try LSTMs today. I'll maybe code it up tomorrow. What I have in mind- A conv layer encoder, flatten, followed by an LSTM layer (why not plain RNNs? well LSTMs handle vanishing gradients, so why the hassle).

Well, what about the previous conv nets?

What I did:

  1. Extensive tuning - of learning rate, batch size, dropout ratio, input size using a grid search
  2. Trained each network for 75 epochs and obtained weights, predicted motion and corresponding FFT, error etc.

What I observed:

  1. Loss curves look okay, validation loss isn't going up, so I don't think overfitting is the issue
  2. Training for over (even) 75 epochs seems to be pointless.

What I think is going wrong:

  1. Input size- relatively large input size: 350 x 350. Here, the input image size seems to be 128 x 128.
  2. Inadequate pre-processing.
    1. I have not applied any filters/blurs etc. to the frames.
    2. I have also not tried dimensionality reduction techniques such as PCA

What I will try now:

  1. Collect new data: with smaller amplitudes and different frequencies
  2. Tune the LSTM network for the data I have
  3. Try new CNN architectures with more aggressive max pooling and fewer parameters
  4. Ensembling the models (see this and this). Right now, I have multiple models trained either with same architecture and different hyperparameters or with different architectures. As a first pass, I intend to average the predictions of all the models and see if that improves performance.
Attachment 1: cnn-lstm.png
cnn-lstm.png
Attachment 2: fft_yaw.pdf
fft_yaw.pdf
Attachment 3: yaw_motion.pdf
yaw_motion.pdf
  14761   Mon Jul 15 14:53:40 2019 MilindUpdateIOOkeyed psl crate, unstick.py

Mode cleaner was not locked today. Koji came in and concluded that PSL had died. So we keyed it. Then we ran my unstick.py code. Mode cleaner is locked now.

Quote:

Today, Gautam keyed the C1PSL crate and we got to test my unstick.py code. It seems to be working fine. Remarks:

  1. Gautam moved the unstick.py code to /opt/rtcds/caltech/c1/scripts/cds. Therefore, the steps to run this code are now:
    1. cd /opt/rtcds/caltech/c1/scripts/cds
    2. python unstick.py c1psl (for the c1psl machine)
  2. There is now a sleepTime global variable in the code which defines the amount of delay between successive channel toggles. We set this to 1ms and it took the code around 3s to run.
  3. Gautam was curious to see if this would work even if we set the sleepTime parameter to 0 but decided that that could be tested the next time something was keyed.
  4. I still need to add the signal handling thing to this code.

Following this, we tested my PMC autolocker code. The code ran for about a minute before achieveing lock. Remarks:

  1. Gautam moved my code (pmc_autolocker.py and autolocker_config.yaml) to /cvs/cds/rtcds/caltech/c1/scripts/PSL/PMC/ . Therefore, the steps to run this code are now:
    1. cd /cvs/cds/rtcds/caltech/c1/scripts/PSL/PMC/
    2. python pmc_autolocker.py (check code or use --help to see what the command line arguments do which is only for when you wanna override the details in the .yaml file)
  2. Gautam suggested that I add some delay between succesive steps of DC output adjust so that it locks quickly. I'll do that ASAP. For now, it works.
  14762   Mon Jul 15 18:55:05 2019 gautamUpdateIOOMegatron hard-rebooted

[koji, gautam]

In addition to c1psl needing a reboot, megatron was un-ssh-able (although it was responding to ping). Clue was that the NPRO PZT control voltage was drifting a lot on the StripTool trace. Koji hard-rebooted the machine. Now IMC is locked, and FSS slow servo is also running.

  14763   Tue Jul 16 15:00:03 2019 gautamUpdateSUSMultiple small EQs

There were several small/medium earthquakes in Ridgecrest and one medium one in Blackhawk CA at about 2000 UTC (i.e. ~ 2 hours ago), one of which caused BS, ITMY, and ETM watchdogs to trip. I restored the damping just now.

  14764   Tue Jul 16 15:17:57 2019 KojiHowToCDSFinal bit bug of the BIO CDS module

Yutaro talked about the BIO bug in KAGRA elog. http://klog.icrr.u-tokyo.ac.jp/osl/?r=9536

I think I made the similar change for the 40m model somewhere (don't remember), but be aware of the presense of this bug.

  14765   Tue Jul 16 16:00:01 2019 gautamUpdateCDSc1iscaux Supermicro setup

I worked on preparing for the c1iscaux upgrade a bit today.

  1. Attachment #1: This shows where the 120 GB solid-state hard-drive and the 2 RAM cards (2GB each) are installed.
    • I found that it required considerable application of force to get the RAM cards into their slots.
    • Note: the 4GB RAM is broken up into two separate physical cards, each 2GB. The labeling is a bit confusing, as each card suggests it is by itself 4GB.
  2. OS install for c1iscaux:
    • I followed Jon's instructions (and added some of mine to the wiki page to hopefully make this process even less thinking-intensive).
    • To be able to use the IP address 192.168.113.83, removed "bscteststand" from chiara martian.hosts and rev.113.168.192.in-addr.arpa as the last mention I could find of this machine was from 2009 (and I'm pretty sure it isn't an active unit anymore). I then restarted the bind9 process. 
    • The hostname for this machine is currently "c1iscaux3" for testing purposes, I will change it once we do the actual install.
    • There was an error in the installation instructions to allow incoming ssh connections - it is openssh-server that is required, not openssh-client. This has now been fixed on the wiki page instructions.
  3. Acromag static IP assignment:
    • Assigned 2 ADCs (XT1221), 5 DACs (XT1541) and 5 sinking BIO units (XT1111) static IP addresses (and labelled them for easy reference) using the windows laptop and the Acromag IP config utility.
    • I saw no reason not to use the 192.168.114.yyy scheme for the Acromag subnet on this machine, even though c1auxex and c1vac both have subnets with this addressing prefix. For reasons unknown to me, Jon opted to use 192.168.115.yyy for the c1susaux Acromag subnet.
  4. Followed the excellent step-by-step to install EPICS, Modbus and Asyn.
    • This took a while, ~1 hour, dominated by the building of EPICS. The other two took only a couple of minutes each.
    • The same combination suggested on Jon's wiki, of Modbus R2-11, EPICS base-7.0.1 and asyn4-33, are the most current at the time of installation.
    • Couple of typos that prevented straight up copy-pasting were fixed on the wiki.
  5. Playground for testing new database files:
    • made a directory /cvs/cds/caltech/target/c1iscaux3 and copied over the .db files from /cvs/cds/caltech/target/c1iscaux and /cvs/cds/caltech/target/c1iscaux2 over.
    • Johannes said he did not develop any code to automate the process of translating the old .db files into the new ones for the Acromag - I won't invest the time in developing any either as I think just manually editing the files will be faster. 
    • I think I will follow the c1susaux convention of grouping .db files by the physical electronics system where possible (e.g. REFL11 channels in one file, CM channels in one file etc), as I think this makes for easier debugging.
    • There is an old "PZT_AI.db" file which I think consists completely of obsolete channels.
  6. Next steps:
    • Wire up the crate [Chub]
    • Make the database files and modbus files for talking to the Acromags on the internal subnet [Gautam], check the .db files [Koji]
    • Wiring of whitening switching from P1 to P2 connector, Issue #1 in this elog (this will also requrie the installation of the DIN shrouds) [Koji]
    • Soldering of P2 interface boards [Gautam]
    • Bench testing [Gautam, Koji, Chub]
    • Installation and in-situ testing [Gautam, Koji, Chub]

All the required additional parts should be here by the end of the week - I'd like to aim for Wednesday 7/24 for the installation in 1Y3 and in-situ testing. While talking to Rana, I realized that we should also factor in the c1aux slow channels into this acromag crate - there is no need for a separate machine to handle the shutters and illuminators. But let's not worry about that for now, those channels can simply be added later.

Attachment 1: IMG_7769.JPG
IMG_7769.JPG
  14766   Wed Jul 17 03:05:01 2019 KruthiUpdateASSMC spot position measurement scripts

[Kruthi, Gautam, Rana]

Gautam installed Atom text editor on Pianosa yesterday.


MC spot position measurement scripts (these can be found in /scripts/ASS/MC directory)

  • Changed the power threshold for MC2 lock loss check from 15000 to 12000 (volts) in the MeasureSpotPositions.py script. This is because, the C1:I00-MC_TRANS_SUM reads a value, usually, greater than 14000 and with 15000 as the threshold, the script will always say the MC isn't locked even though it is!. Also, to account for additional variation we have a margin of 2000.
  • Issues with datetime: though MeasureSpotPositions.py was creating a .dat file, MC_spotMeasurement_history.py threw an error because the .dat file's name was not in the required format. I fixed this bug.
  • Just running the MeasureSpotPositions.py doesn't enter the results into the log file, instead ./mcassMCdecenter should be run
  • MC_spotMeasurement_history.py just plots the spot positions (in mm) vs days since 2013, using the log file. It still has some bugs
  14767   Wed Jul 17 17:56:18 2019 KojiConfigurationComputersGave resolv.conf to giada

Kruthi noticed that she could not login to rossa from giada.

I checked /etc/resolv.conf and it was

nameserver 127.0.0.1

so obviously it is useless to refer localhost (i.e. giada) as a nameserver.

I copied our usual resolv.conf to giada as following:

nameserver 192.168.113.104
nameserver 131.215.125.1
nameserver 8.8.8.8

search martian

Giada's ssh known_host had unupdated entry for rossa, so I had to clean it up, but after that we can connect to rossa from giada just by "ssh rossa".

Case closed.

  14768   Wed Jul 17 20:12:26 2019 KruthiUpdateCamerasAnother GigE in place of analog camera

I've taken the MC2 analog camera down and put another GigE (unit 151) in its place. This is just temporary and I'll put the analog camera back once I finish the MC2 loss map calibration. I'm using a 25mm focal length camera lens with it and it gives a view of MC2 similar to the analog camera one. But I don't think it is completely focused yet (pictures attached).

...more to follow

gautam - Attachment #3 is my (sad) attempt at finding some point scatterers - Kruthi is going to play around with photUtils to figure out the average size of some point scatterers.

Attachment 1: zoomed_out_gige.png
zoomed_out_gige.png
Attachment 2: osems_mc2.png
osems_mc2.png
Attachment 3: MC2.pdf
MC2.pdf
  14769   Wed Jul 17 21:22:41 2019 gautamUpdateCDSCM board Latch Enable subtlety

[koji, gautam]

Koji pointed out an important subtlety pertaining to the "LATCH ENABLE" signal line on the CM board. The purpose of this line is to smoothly facilitate the transition of a change in the "multi-bit-binary-outputs", a.k.a. "mbbo", that are controlled by MEDM gain sliders, to the analog electronics on the CM board. Why is this necessary? Imagine changing the gain from 7dB (=0111 in mbbo representation) to 8dB (=1000 in mbbo representation). In order to realize this change, all 4 bits have to change their state. But this almost certainly doesn't happen synchronously, because our EPICS interface isn't synchronous. So at some intermediate times, the mbbo representation could be 0100 (=4dB), or 1111 (=15dB), or many other possible values, which are all significantly different from either the initial value or the desired final state. This is clearly undesirable.

In order to protect against this kind of error, a Latched output part, 74ALS573, is used to buffer the physical digital logic levels from the switches in the analog gain stages. So in the default state, the "LATCH ENABLE" signal line is held "LOW". When a change happens in the EPICS value corresponding to a gain slider, the "LATCH ENABLE" state is quickly toggled to "HIGH", so as to enable the appropriate analog gain stages to be switched, and then again to "LOW", at which point the latch holds its output state. This logic is currently implemented by a piece of code called "latch.o", which is the compiled version of "latch.st", which may be found in /cvs/cds/caltech/target/c1iool0 where it presumably was written for the IMC servo board, but not in /cvs/cds/caltech/target/c1iool0  , which is where the CM board database files reside. The only elog reference I can find pertaining to this particular piece of code is from Alan, and doesn't say anything about the actual logic.

For the new c1iscaux, we need to implement this logic somehow. After discussion between Koji and me, we feel that a piece of python code is sufficient. This would continuously run in the background on the supermicro server machine. The channel hierarchy for each gain channes is as follows (I've taken the example of C1:LSC-CM_REFL1_GAIN):

  • C1:LSC-CM_REFL1_GAIN ------ this is the channel tied to an MEDM slider, and so is a "soft" channel
  • C1:LSC-CM_REFL1_SET ------- this is a "soft" channel that gets converted to an mbbo
  • C1:LSC-CM_REFL1_BITS ------ this is a channel that actually controls (multiple) physical binary outputs on the Acromag

So the logic will be that it continuously scans the EPICS channel C1:LSC-CM_REFL1_GAIN  for a change in set value. When a change is detected, it has to update the C1:LSC-CM_REFL1_SET channel. In the next EPICS refresh cycle, this would result in the mbbo bits, C1:LSC-CM_REFL1_BITS , all changing to the appropriate values. After these changes have happened, we need to toggle the LATCH ENABLE in order to allow the changes to propagate to the analog gain stage switches. Need to think about what's the best way to do this.

  14770   Thu Jul 18 00:51:52 2019 KojiSummaryCDSiscaux electronics modifications

Along with the plan in ELOG 14744, the ISC PD interface and the whitening filter board have been modificed. The ISC PD I/Fs were restored to the crate and the cables were connected. The whitening filteres are still on the electronics bench for some more tests before being returned to the crate.

The updated schematics were uploaded as https://dcc.ligo.org/D1900318 and https://dcc.ligo.org/D1900319

- Modification of the ISC PD interface: Jumpers between DIN96 P1 and P2. Replace all AD797s with OP27. In fact only I/F #1 (the left most)  had total 12 AD797 but the other units already had OP27s.

- Modification of the whitening filter: Jumpers between DIN96 P1 and P2.

Attachment 1: LSC_whitening2.jpg
LSC_whitening2.jpg
Attachment 2: LSC_whitening.jpg
LSC_whitening.jpg
  14771   Thu Jul 18 10:46:04 2019 gautamUpdateCDSDatabase files made

I completed the translation of the .db files for the EPICS database records from the VME notation to the Acromag/Modbus/Asyn notation. The channels are now organized into 5 database files, located in /cvs/cds/caltech/target/c1iscaux3/,  for convenience:

  1. C1_ISC-AUX_LSCPDs.db -------- This handles whitening gain, AA enable/bypass, Demodulator FE, and PD Interface Board channels for REFL11, REFL55, REFL33, REFL165, POP22, POP110, POX11, POY11, AS55 and AS110 photodiodes.
  2. C1_ISC-AUX_CM.db -------------- This handles all channels for the CM board. The mbbo addressing notation needs to be checked.
  3. C1_ISC-AUX_QPDs.db ----------- This handles all channels for the IPPOS QPD.
  4. C1_ISC-AUX_ALS.db ------------- This handles all channels for the IR ALS DFD LO and RF power monitoring.
  5. C1_ISC-AUX_SPARE.db ---------- This handles the unused channels for the various whitening, AA and PD interface boards.

For reasons unknown to me, the database files in the other Acromag system target directories (e.g. c1susaux, c1auxex) all had 755 level access permission - maybe this is required for systemctl to handle the EPICS serving? Anyways, I upgraded the permission level of the above 5 files using chmod.

There are almost certainly typos / other errors, and I may have missed copying over some soft/calibrated channels, but I hope that this way of grouping by subsystem will make the debugging less painful. Once Chub connects up the power lines to the Acromags, I will run the soft tests. For this purpose, I've also made a C1_ISC-AUX.cmd file and a C1_ISC-AUX.env file in the above target directory, and also made the modbusIOC.service file in /etc/systemd/system on the supermicro.

  14773   Thu Jul 18 19:58:56 2019 gautamUpdateCDSWork on Acromag chassis

Now that the .db files were prepared, I wanted to test for errors. So I did the following:

  1. Acromags were mounted on the DIN rails. Attachment #1 shows the grouping of ADC, DAC and BIO units. They are labelled with their IP addresses.
  2. Wiring of power:
    • Chub had already prepared the backplane with the power connectors, switches and indicator LEDs.
    • So I just had to daisy chain the +24 V (RED) and GND (BLACK) terminals for all the acromags together, which I did using 24 AWG wire (we may want to use heavier gauge given the current draw).
  3. Ethernet cables were used to daisy chain the network connectivity between the various units.  Attachment #1 shows the current state of the chassis box.
  4. Front panel pieces were attached and labelled, see Attachment #2
    • I found it was sufficient to use the front - we may use the rear panel slots when we want to add connections for controlling the c1aux machine channels.
    • The D15 P2 connector panel for the CM board will arrive tomorrow and will be installed then.
  5. Entire setup was connected to power and ethernet, see Attachment #3
    • As usual, the current draw is significant for the collection of Acromags, I got around this problem by using the bench supply to "Parallel" mode to enhance the current driving capacity.
    • For the ethernet connection, I used the office space port #6, which I connected at the network rack end to the eth1 port of the Supermicro.

All the Acromags are seen on the 192.168.114 subnet on c1iscaux3 yes- however, when I run the modbusIOC process, I see various errors in the logfile no, so more debugging is required. Nevertheless, progress.

Update 2245: Turns out the errors were indeed due to a copy/paste error - I had changed the IP addresses for the ADCs from the .115 subnet c1susaux was using, but forgot to do so for the DACs and BIOs. Now, if I turn off the existing c1iscaux so that there aren't any EPICS clashes, the EPICS server initializes correctly. There are still some errors in the log file - these pertain to (i) the mbbo notation, which I have to figure out, and (ii) the fact that this version of EPICS, 7.0.1, does not support channel descriptions longer than 28 characters (we have several that exceed this threshold). I think the latter isn't a serious problem.

Getting closer... Note that I turned off the c1iscaux VME crate to prevent any EPICS server clashes. I will turn it back on tomorrow.

Attachment 1: IMG_7771.JPG
IMG_7771.JPG
Attachment 2: IMG_7770.JPG
IMG_7770.JPG
Attachment 3: IMG_7772.JPG
IMG_7772.JPG