  11027   Sat Feb 14 00:42:02 2015 KojiUpdateGeneralRF amplifier for ALS

The RF analyzer was returned to the control room. There are two beat notes from X/Y confirmed.

I locked the arms and aligned them with ASS.

When the end greens are locked at TEM00, X/Y beat amplitudes were ~33dBm and ~17dBm. respectively.
I don't judge if they are OK or not, as I don't recall the nominal values.

  11026   Fri Feb 13 19:41:13 2015 ericqUpdateGeneralRF amplifier for ALS

As Koji found one of the spare channels of the ALS/FOL RF amplifier box nonfunctional yesterday, I pulled it out to fix it. I found that one of the sma cables did not conduct.

It was replaced with a new cable from Koji. Also, I rearranged the ports to be consistent across the box, and re-labeled with the gains I observed. 

It has been reinstalled, and the Y frequency counter that is using one of the channels shows a steady beat freq.

I cannot test the amplitude of the green X beat at this time, as Koji is on the PSL table with the PSL shutter closed, and is using the control room spectrum analyzer. However, the dataviewer trace for the fine_phase_out_Hz looks like free swinging cavity motion, so its probably ok.

  11025   Fri Feb 13 18:56:44 2015 ranaUpdateElectronicsSecond QPD Whitening Switch enabled

Depends on the plots of the whitening I guess; if its low freq sat, then we lower the light level with ND filters. If its happening above 10 Hz, then we switch off the whitening.

Quote:

Went to zero CARM offset on ALS; transmission QPDs are still saturating :(

Maybe we need to switch off all whitening.

 

  11024   Fri Feb 13 17:07:51 2015 JenneUpdateElectronicsSecond QPD Whitening Switch enabled

I first updated the DCC branches for the Xend and Yend to reflect the as-built situation from December 2014, and then I updated the drawings after Q's modifications today.

  11023   Fri Feb 13 14:59:13 2015 ericqUpdateElectronicsSecond QPD Whitening Switch enabled

Went to zero CARM offset on ALS; transmission QPDs are still saturating :(

Maybe we need to switch off all whitening.

  11022   Fri Feb 13 14:27:30 2015 ericqUpdateElectronicsSecond QPD Whitening Switch enabled

I have re-enabled the second whitening stage switching on each quadrant of each end's QPD whitening board, to try and avoid saturations at full power. Looking at the spectra while single arm locked, I confirmed that the FM2 whitening switch works as expected. FM1 should be left on, as it is still hard-wired to whiten. 

The oscillations in the Y QPD still exist. Jenne is updating the schematics on the DCC.

  11021   Fri Feb 13 03:44:56 2015 JenneUpdateLSCHeld using ALS for a while at "0" CARM offset with PRMI

[Jenne, Rana]

We wanted to jump right in and see if we were ready to try the new "ALS fool" loop decoupling scheme, so we spent some time with CARM and DARM at "0" offset, held on ALS, with PRMI locked on REFL33I&Q (no offsets).  Spoiler alert:  we weren't ready for the jump.

The REFL11 and AS55 PDs had 0dB analog whitening, which means that we weren't well-matching our noise levels between the PD noise and the ADC noise.  The photodiodes have something of the order nanovolt level noise, while the ADC has something of the order microvolt level noise.  So, we expect to need an analog gain of 1000 somewhere, to make these match up.  Anyhow, we have set both REFL11 and AS55 to 18dB gain. 

On a related note, it seems not so great for the POX and POY ADC channels to be constantly saturated when we have some recycling gain, so we turned their analog gains down from 45dB to 0dB.  After we finished with full IFO locking, they were returned to their nominal 45dB levels. 

We also checked the REFL33 demod phase at a variety of CARM offsets, and we see that perhaps it changes by one or two degrees for optimal rotation, but it's not changing drastically.  So, we can set the REFL33 demod phase at large CARM offset, and trust it at small CARM offset.

We then had a look at the transmon QPD inputs (before the dewhitening) for each quadrant.  They are super-duper saturating, which is not so excellent. 
QPDsaturation_12Feb2015.pdf

We think that we want to undo the permanently-on whitening situation.  We want to make the second stage of whitening back to being switchable.  This means taking out the little u-shaped wires that are pulling the logic input of the switches to ground.  We think that we should be okay with one always on, and one switchable.  After the modification, we must check to make sure that the switching behaves as expected.  Also, I need to figure out what the current situation is for the end QPDs, and make sure that the DCC document tree matches reality.  In particular, the Yend DCC leaf doesn't include the gain changes, and the Xend leaf which does show those changes has the wrong value for the gain resistor.

After this, we started re-looking at the single arm cancellation, as Rana elogged about separately.

ALSfool_12Feb2015.pdf

Attachment 1: QPDsaturation_12Feb2015.pdf
QPDsaturation_12Feb2015.pdf
Attachment 2: ALSfool_12Feb2015.pdf
ALSfool_12Feb2015.pdf
  11020   Fri Feb 13 03:28:34 2015 ranaUpdateLSCNew Locking Paradigm - Loop-gebra

Not so fast!

In the drawing, the FF path should actually be summed in after the Phase Tracker (i.e. after S_ALS). This means that the slow response of the phase tracker needs to be taken into account in the FF cancellation filter. i.e. D = -A_REFL * P_ALS * S_ALS. Since the Phase Tracker is a 1/f loop with a 1 kHz UGF, at 100 Hz, we can only get a cancellation factor of ~10.

So, tonight we added a 666:55 boost filter into the phase tracker filter bank. I think this might even make the ALS locking loops less laggy. The boost is made to give us better tracking below ~200 Hz where we want better phase performance in the ALS and more cancellation of the ALS-Fool. If it seems to work out well we can keep it. If it makes ALS more buggy, we can just shut it off.

Its time to take this loop cartoon into OmniGraffle.

  11019   Thu Feb 12 23:47:45 2015 KojiUpdateLSC3f modulation cancellation

- I built another beat setup on the PSL table at the South East side of the table.
- The main beam is not touched, no RF signal is touched, but recognize that I was present at the PSL table.
- The beat note is found. The 3rd order sideband was not seen so far.
- A PLL will be built tomorrow. The amplifier box Manasa made will be inspected tomorrow.

- One of the two beams from the picked-off beam from the main beam line was introduced to the beat setup.
(The other beam is used of for the beam pointing monitors)
There is another laser at that corner and the output from this beam is introduced into the beat setup.
The combined beam is introduced to PDA10CF (~150MHz BW).

- The matching of the beam there is poor. But without much effort I found the beat note.
  The PSL laser had 31.33 deg Xtal temp. When the beat was found, the aux laser had the Xtal temp of 40.88.

- I could observe the sidebands easily, with a narrower BW of the RF analizer I could see the sidebands up to the 2nd order.
  The 3rd order was not seen at all.

- The beat note had the amplitude of about -30dBm. One possibility is to amplify the signal. I wanted to use a spare channel
of the ALS/FOLL amplifier box. But it gave me rather attenuation than any amplification.
I'll look at the box tomorrow.

- Also the matching of two beams are not great. The PD also has clipping I guess. These will also be improved tomorrow

- Then the beat note will be locked at a certain frequency using PLL so that we can reduce the measurement BW more.

 

  11018   Thu Feb 12 23:40:12 2015 ranaUpdateIOOBounce / Roll stopband filters added to MC ASC filter banks

The filters were already in the damping loops but missing the MC WFS path. I checked that these accurately cover the peaks at 16.5 Hz and 23.90 and 24.06 Hz.

Attachment 1: 59.png
59.png
  11017   Thu Feb 12 22:28:16 2015 JenneUpdateLSCNew Locking Paradigm - LSC model changes, screens modified

I have modified the LSC trigger matrix screen, as well as the LSC overview screen, to reflect the modifications to the model from yesterday. 

Also, I decided that we probably won't ever want to trigger the zero crossing on the Q phase signals of REFL.  Instead, we may want to try it out with the single arms, so the zero crossing selection matrix is now REFL11I, REFL55I, POX11I, POY11I, in that order. 

 

  11016   Thu Feb 12 19:18:49 2015 JenneUpdateLSCNew Locking Paradigm - Loop-gebra

EDIT, JCD, 17Feb2015:  Updated loop diagram and calculation: http://131.215.115.52:8080/40m/11043


Okay, Koji and I talked (after he talked to Rana), and I re-looked at the original cartoon from when Rana and I were thinking about this the other day.

The original idea was to be able to actuate on the MC frequency (using REFL as the sensor), without affecting the ALS loop.  Since actuating on the MC will move the PSL frequency around, we need to tell the ALS error signal how much the PSL moved in order to subtract away this effect. (In reality, it doesn't matter if we're actuating on the MC or the ETMs, but it's easier for me to think about this way around).  This means that we want to be able to actuate from point 10 in the diagram, and not feel anything at point 4 in the diagram (diagram from http://131.215.115.52:8080/40m/11011)

This is the same as saying that we wanted the green trace in http://131.215.115.52:8080/40m/11009 to be zero.

So.  What is the total TF from 10 to 4? 

{\rm TF}_{\rm (10 \ to \ 4)} = \frac{D_{\rm cpl} + {\color{DarkRed} A_{\rm refl}} {\color{DarkGreen} P_{\rm als}}}{1-{\color{DarkRed} A_{\rm refl} G_{\rm refl} S_{\rm refl} P_{\rm refl}} - {\color{DarkGreen} A_{\rm als} G_{\rm als} S_{\rm als}} ({\color{DarkGreen} P_{\rm als}} + D_{\rm cpl} {\color{DarkRed} G_{\rm refl} P_{\rm refl} S_{\rm refl}})}

So, to set this equal to zero (ALS is immune to any REFL loop actuation), we need D_{\rm cpl} = - {\color{DarkRed} A_{\rm refl}} {\color{DarkGreen} P_{\rm als}}.

Next up, we want to see what this means for the closed loop gain of the whole system.  For simplicity, let's let H_* = A_* G_* S_* P_*, where * can be either REFL or ALS. 

Recall that the closed loop gain of the system (from point 1 to point 2)  is

{\rm TF}_{\rm (1 \ to \ 2)} = \frac{1}{1-{\color{DarkRed} A_{\rm refl} G_{\rm refl} S_{\rm refl} P_{\rm refl}} - {\color{DarkGreen} A_{\rm als} G_{\rm als} S_{\rm als}} ({\color{DarkGreen} P_{\rm als}} + D_{\rm cpl} {\color{DarkRed} G_{\rm refl} P_{\rm refl} S_{\rm refl}})} , so if we let  D_{\rm cpl} = - {\color{DarkRed} A_{\rm refl}} {\color{DarkGreen} P_{\rm als}} and simplify, we get

{\rm TF}_{\rm (1 \ to \ 2)} = \frac{1}{1-{\color{DarkRed} H_{\rm refl}} - {\color{DarkGreen} H_{\rm als}} + {\color{DarkRed} H_{\rm refl}}{\color{DarkGreen} H_{\rm als}}}

This seems a little scary, in that maybe we have to be careful about keeping the system stable.  Hmmmm.  Note to self:  more brain energy here.


Also, this means that I cannot explain why the filter wasn't working last night, with the guess of a complex pole pair at 1Hz for the MC actuator.  The  ALS plant has a cavity pole at ~80kHz, so for our purposes is totally flat.  The only other thing that comes to mind is the delays that exist because the ALS signals have to hop from computer to computer.  But, as Rana points out, this isn't really all that much phase delay below 100Hz where we want the cancellation to be awesome. 

I propose that we just measure and vectfit the transfer function that we need, since that seems less time consuming than iteratively tweaking and checking. 

Also, I just now looked at the wiki, and the MC2 suspension resonance for pos is at 0.97Hz, although I don't suspect that that will have changed anything significantly above a few Hz.  Maybe it makes the cancellation right near 1Hz a little worse, but not well above the resonance.

 

  11015   Thu Feb 12 15:21:37 2015 ericqUpdateComputer Scripts / Programsnetgpib updates

I've fixed the gpib scripts for the SR785 and AG4395A to output data in the same format as expected by older scripts when called by them. In addition, there are now some easier modes of operation through the measurement scripts SRmeasure and AGmeasure. These are on the $PATH for the main control room machines, and live in scripts/general/netgpib


Case 1: I manually set up a measurement on the analyzer, and just want to download / plot the data. 

Make sure you have a yellow prologix box plugged in, and can ping the address it is labeled with. (i.e. 'vanna'). Then, in the directory you want to save the data, run:

SRmeasure -i vanna -f mydata --getdata --plot

This saves mydata_(datetime).txt and mydata_(datetime).pdf in the current directory. 

In all cases, AGmeasure has the identical syntax. If the GPIB address is something other than 10, specifiy it with -a, but this is rarely the case.


Case 2: I want to remotely specify a measurement

Rather than a series of command line arguments, which may get lost to the mists of time, I've set the scripts up to use parameter files that serve as arguments to the scripts. 

Get the templates for spectrum and TF measurements in your current directory by running

SRmeasure --template

Set the parameters with your text editor of choice, such as frequency span, filename output, whether to create a plot or not, then run the measurement:

SRmeasure SR785template.yml


Case 3: I want to compare my data with previous measurements

In the template parameter files, there is an option 'plotRefs', that will automatically plot the data from files whose filenames start with the same string as the current measurement.

If, in the "#" commented out header of the data file, there is a line that contains "memo:" or "timestamp:", it will include the text that follows in the plot legend. 


There are also methods to remotely trigger an already configured measurement, or remotely reset an unresponsive instrument. Options can be perused by looking at the help in SRmeasure -h

I've tested, debugged, and used them for a bit, but wrinkles may remain. They've been svn40m committed, and I also set up a separate git repository for them at github.com/e-q/netgpibdata

  11014   Thu Feb 12 12:23:21 2015 manasaUpdateGeneralWaking up the PDFR measurement system

[EricG, Manasa]

We woke up the PDFR measurement setup that has been sleeping since summer. We ran a check for the laser module and the multiplexer module. We tried setting things up for measuring frequency response of AS55.
We could not repeat Nichin's measurements because the gpib scripts are outdated and need to be revised. 

PDFR diode laser was shutdown after this job.

  11013   Thu Feb 12 12:16:04 2015 manasaUpdateGeneralFiber shielding

[Steve, Manasa]

The fibers around the PSL table were shielded to avoid any tampering.

  11012   Thu Feb 12 11:59:58 2015 KojiUpdateLSCNew Locking Paradigm - Loop-gebra

The goals are:

- When the REFL path is dead (e.g. S_REFL = 0), the system goes back to the ordinary ALS loop. => True (Good)

- When the REFL path is working, the system becomes insensityve to the ALS loop
(i.e. The ALS loop is inactivated without turning off the loop.) => True when (...) = 0

Are they correct?

 

Then I just repeat the same question as yesterday:

S is a constant, and Ps are cavity poles. So,  approximately to say, (...) = 0 is realized by making D = 1/G_REFL.
In fact, if we tap the D-path before the G_REFL, we remove this G_REFL from (...). (=simpler)
But then, this means that the method is rather cancellation between the error signals than
cancellation between the actuation. Is this intuitively reasonable? Or my goal above is wrong?

  11011   Thu Feb 12 11:14:29 2015 JenneUpdateLSCNew Locking Paradigm - Loop-gebra

I have calculated the response of this new 2.5 loop system.

The first attachment is my block diagram of the system.  In the bottom left corner are the one-hop responses from each green-colored point to the next.  I use the same matrix formalism that we use for Optickle, which Rana described in the loop-ology context in http://nodus.ligo.caltech.edu:8080/40m/10899

In the bottom right corner is the closed loop response of the whole system.

 

Also attached is a zipped version of the mathematica notebook used to do the calculation.

EDIT, JCD, 17Feb2015:  Updated loop diagram and calculation:  http://131.215.115.52:8080/40m/11043

Attachment 1: ALS_REFL_comboLockingCartoon_11Feb2015.PDF
ALS_REFL_comboLockingCartoon_11Feb2015.PDF
Attachment 2: ALS_REFL_comboLocking_11Feb2015.zip
  11010   Thu Feb 12 03:43:54 2015 ericqUpdateLSC3F PRMI at zero ALS CARM

I have been able to recover the ability to sit at zero CARM offset while the PRMI is locked on RELF33 and CARM/DARM are on ALS, effectively indefinitely. However, I feel like the transmon QPDs are not behaving ideally, because the reported arm powers freqently go negative as the interferometer is "buzzing" through resonance, so I'm not sure how useful they'll be as normalizing signals for REFL11. I tried tweaking the DARM offset to help the buildup, since ALS is only roughly centered on zero for both CARM and DARM, but didn't have much luck.

Example:

Turning off the whitening on the QPD segments seems to make everything saturate, so some thinking with daytime brain is in order.


How I got there:

It turns out triggering is more important than the phase margin story I had been telling myself. Also, I lost a lot of time to needing demod angle change in REFL33. Maybe I somehow caused this when I was all up on the LSC rack today?

We have previously put TRX and TRY triggering elements into the PRCL and MICH rows, to guard against temporary POP22 dips, because if arm powers are greater than 1, power recylcing is happening, so we should keep the loops engaged. However, since TRX and TRY are going negative when we buzz back and forth through the resonsnace, the trigger row sums to a negative value, and the PRMI loops give up. 

Instead, we can used the fortuitously unwhitened POPDC, which can serve the same function, and does not have the tendancy to go negative. Once I enabled this, I was able to just sit there as the IFO angrily buzzed at me. 

Here are my PRMI settings

REFL33 - Rotation 140.2 Degrees, -89.794 measured diff

PRCL = 1 x REFL33 I; G = -0.03; Acquire FMs 4,5; Trigger FMs 2, 9; Limit: 15k ; Acutate 1 x PRM

MICH = 1 x REFL33 Q, G= 3.0, Acquire FMs 4,5,8; Trigger FM 2, 3; Limit: 30k; Actuate -0.2625 x PRM + 0.5 x BS

Triggers = 1 x POP22 I + 0.1 * POPDC, 50 up 5 down


Just for kicks, here's a video of the buzzing as experienced in the control room

Attachment 1: Feb12_negativeTR.png
Feb12_negativeTR.png
  11009   Thu Feb 12 01:43:09 2015 ranaUpdateLSCNew Locking Paradigm - LSC model changes

With the Y Arm locked, we checked that we indeed can get loop decoupling using this technique.

The guess filter that we plugged in is a complex pole pair at 1 Hz. We guessed that the DC gain should be ~4.5 nm count. We then converted this number into Hz and then into deg(?) using some of Jenne's secret numbers. Then after measuring, we had to increase this number by 14.3 dB to an overall filter module gain of +9.3.

The RED trace is the usual 'open loop gain' measurement we make, but this time just on the LSC-MC path (which is the POY11_I -> ETMY path).

The BLUE trace is the TF between the ALS-Y phase tracker output and the FF cancellation signal. We want this to be equal ideally.

The GREEN trace is after the summing point of the ALS and the FF. So this would go to zero when the cancellation is perfect.

So, not bad for a first try. Looks like its good at DC and worse near the red loop UGF. It doesn't change much if I turn off the ALS loop (which I was running with ~10-15x lower than nominal gain just to keep it out of the picture). We need Jenne to think about the loop algebra a little more and give us our next filter shape iteration and then we should be good.

Attachment 1: TF.gif
TF.gif TF.gif TF.gif TF.gif TF.gif TF.gif TF.gif TF.gif TF.gif TF.gif TF.gif TF.gif TF.gif TF.gif TF.gif TF.gif TF.gif TF.gif TF.gif TF.gif TF.gif TF.gif TF.gif TF.gif TF.gif TF.gif
  11008   Thu Feb 12 01:00:18 2015 ranaUpdateLSCRFPD spectra

The nonlinearity in the LSC detection chain (cf T050268) comes from the photodetector and not the demod board. The demod board has low pass or band pass filters which Suresh installed a long time ago (we should check out what's in REFL33 demod board). 

Inside the photodetector the nonlinearity comes about because of photodiode bias modulation (aka the Grote effect) and slew rate limited distortion in the MAX4107 preamp.

  11007   Wed Feb 11 22:13:44 2015 JenneUpdateLSCNew Locking Paradigm - LSC model changes

In order to try out the new locking scheme tonight, I have modified the LSC model.  Screens have not yet been made.

It's a bit of a special case, so you must use the appropriate filter banks:

CARM filter bank should be used for ALS lock.  MC filter bank should be used for the REFL1f signal. 

The output of the MC filter bank is fed to a new filter bank (C1:LSC-MC_CTRL_FF).  The output of this new filter bank is summed with the error point of the CARM filter bank (after the CARM triggered switch).

The MC triggering situation is now a little more sophisticated than it was.  The old trigger is still there (which will be used for something like indicating when the REFL DC has dipped).  That trigger is now AND-ed with a new zero crossing trigger, to make the final trigger decision.  For the zero crossing triggering, there is a small matrix (C1:LSC-ZERO_CROSS_MTRX) to choose what REFL 1f signal you'd like to use (in order, REFL11I, REFL11Q, REFL55I, REFL55Q).  The absolute value of this is compared to a threshold, which is set with the epics value C1:LSC-ZERO_CROSS_THRESH.  So, if the absolute value of your chosen RF signal is lower than the threshold, this outputs a 1, which is AND-ed by the usual schmidt trigger. 

At this moment, the input and output switches of the new filter bank are off, and the gain is set to zero.  Also, the zero crossing selection matrix is all zeros, and the threshold is set to 1e9, so it is always triggered, which means that effectively MC filter bank just has it's usual, old triggering situation.

  11006   Wed Feb 11 18:44:53 2015 manasaUpdateGeneralOptical fiber module

I have moved the optical fiber module for FOL to the PSL table. It is setup on the optical table right now for testing.

Once tests are done, the box will move to the rack inside the PSL enclosure. 

While doing any beat note alignment, please watch out for the loose fibers at the north side of the PSL enclosure until they are sheilded securely (probably tomorrow morning).

  11005   Wed Feb 11 18:11:46 2015 KojiSummaryLSC3f modulation cancellation

33MHz sidebands can be elliminated by careful choice of the modulation depths and the relative phase between the modulation signals.
If this condition is realized, the REFL33 signals will have even more immunity to the arm cavity signals because the carrier signal will lose
its counterpart to produce the signal at 33MHz.

Formulation of double phase modulation

m1: modulation depth of the f1 modulation
m2: modulation depth of the f2 (=5xf1) modulation

The electric field of the beam after the EOM

E=E_0 \exp \left[ {\rm i} \Omega t + m_1 \cos \omega t +m_2 \cos 5 \omega t \right ]
\flushleft = {\it E}_0 e^{{\rm i} \Omega t} \\ \times \left[ J_0(m_1) + J_1(m_1) e^{{\rm i} \omega t}- J_1(m_1) e^{-{\rm i} \omega t} + J_2(m_1) e^{{\rm i} 2\omega t}+ J_2(m_1) e^{-{\rm i} 2\omega t} + J_3(m_1) e^{{\rm i} 3\omega t}- J_3(m_1) e^{-{\rm i} 3\omega t} + \cdots \right] \\ \times \left[ J_0(m_2) + J_1(m_2) e^{{\rm i} 5 \omega t}- J_1(m_2) e^{-{\rm i} 5 \omega t} + \cdots \right]
\flushleft = {\it E}_0 e^{{\rm i} \Omega t} \\ \times \left\{ \cdots + \left[ J_3(m_1) J_0(m_2) + J_2(m_1) J_1(m_2) \right] e^{{\rm i} 3 \omega t} - \left[ J_3(m_1) J_0(m_2) + J_2 (m_1) J_1(m_2) \right] e^{-{\rm i} 3 \omega t} + \cdots \right\}

Therefore what we want to realize is the following "extinction" condition
J_3(m_1) J_0(m_2) + J_2(m_1) J_1(m_2) = 0

We are in the small modulation regime. i.e. J0(m) = 1, J1(m) = m/2, J2(m) = m2/8, J3(m) = m3/48
Therefore we can simplify the above exitinction condition as

m_1 + 3 m_2 = 0

m2 < 0 means the start phase of the m2 modulation needs to be 180deg off from the phase of the m1 modulation.

E = E_0 \exp\left\{ {\rm i} [\Omega t + m_1 \cos \omega t + \frac{m_1}{3} \cos (5 \omega t + \pi)] \right \}

Field amplitude m1=0.3, m2=-0.1 m1=0.2, m2=0.2
Carrier 0.975 0.980
1st order sidebands 0.148 9.9e-2
2nd 1.1e-3 4.9e-3
3rd 3.5e-7 6.6e-4
4th 7.4e-3 9.9e-3
5th 4.9e-2 9.9e-2
6th 7.4e-3 9.9e-3
7th 5.6e-4 4.9e-4
8th 1.4e-5 4.1e-5
9th 1.9e-4 5.0e-4
10th 1.2e-3 4.9e-3
11th 1.9e-4 5.0e-4
12th 1.4e-5 2.5e-5
13th 4.7e-7 1.7e-6
14th 3.1e-6 1.7e-5
15th 2.0e-5 1.6e-4

 

  11004   Wed Feb 11 18:07:42 2015 ericqUpdateLSCRFPD spectra

After some discussion with Koji, I've asked Steve to order some SBP-30+ bandpass filters as a quick and cheap way to help out REFL33. (Also some SBP-60+ for 55MHz, since we only have 1*fmod and 2*fmod bandpasses here in the lab). 

  11003   Wed Feb 11 17:31:11 2015 ericqUpdateLSCRFPD spectra

For future reference, I've taken spectra of our various RFPDs while the PRMI was sideband locked on REFL33, using a 20dB RF coupler at the RF input of the demodulator boards. The 20dB coupling loss has been added back in on the plots. Data files are attached in a zip.

Exceptions: 

  • The REFL165 trace was taken at the input of the amplifier that immediately preceeds the demod board. 
  • The 'POPBB' trace was taken with the coupler at the input of the bias tee, that leads to an amplifier, then splitter, then the 110 and 22 demod boards. 

I also completely removed the cabling for REFLDC -> CM board, since it doesn't look like we plan on using it anytime in the immediate future. 

Attachment 1: REFL.png
REFL.png
Attachment 2: AS.png
AS.png
Attachment 3: POP.png
POP.png
Attachment 4: 2015-02-PDSpectra.zip
  11002   Wed Feb 11 16:49:41 2015 ranaUpdateelogELOGD restarted

No elog response from outside and no elogd process frownon nodus, so I restarted it using 'start-elog.csh'.

  11001   Wed Feb 11 04:08:53 2015 JenneUpdateLSCNew Locking Paradigm?

[Rana, Jenne]

While meditating over what to do about the fact that we can't seem to hold PRMI lock while reducing the CARM offset, we have started to nucleate a different idea for locking

We aren't sure if perhaps there is some obvious flaw (other than it may be tricky to implement) that we're not thinking about, so we invite comments.  I'll make a cartoon and post it tomorrow, but the idea goes like this.....

Can we use ALS to hold both CARM and DARM by actuating on the ETMs, and sit at (nominally) zero offset for all degrees of freedom?  PRMI would need to be stably held with 3f signals throughout this process. 

1) Once we're close to zero offset, we should see some PDH signal in REFL11.  With appropriate triggering (REFLDC goes low, and REFL11I crosses zero), catch the zero crossing of REFL11I, and feed it back to MC2. We may want to use REFL11 normalized by the sum of the arm transmissions to some power (1, 0.5, or somewhere in between may maximize the linear range even more, according to Kiwamu).  The idea (very similar to the philosophy of CESAR) is that we're using ALS to start the stabilization, so that we can catch the REFL11 zero crossing. 

2) Now, the problem with doing the above is that actuating on the mode cleaner length will change the laser frequency.  But, we know how much we are actuating, so we can feed forward the control signal from the REFL11 carm loop to the ALS carm loop.  The goal is to change the laser frequency to lock it to the arms, without affecting the ALS lock.  This is the part where we assume we might be sleepy, and missing out on some obvious reason why this won't work.

3) Once we have CARM doubly locked (ALS pushing on ETMs, REFL11 pushing on MC/laser frequency), we can turn off the ALS system. Once we have the linear REFL11 error signal, we know that we have enough digital gain and bandwidth to hold CARM locked, and we should be able to eek out a slightly higher UGF since there won't be as many digital hops for the error signal to transverse. 

4) The next step is to turn on the high bandwidth common mode servo.  If ALS is still on at this point, it will get drowned out by the high gain CM servo, so it will be effectively off. 

5) Somewhere in here we need to transition DARM to AS55Q.  Probably that can happen after we've turned on the digital REFL11 path, but it can also probably wait until after the CM board is on.

The potential show-stoppers:

Are we double counting frequency cancellation or something somewhere?  Is it actually possible to change the laser frequency without affecting the ALS system?

Can we hold PRMI lock on 3f even at zero CARM offset?  Anecdotally from a few trials in the last hour or so, it seems like coming in from negative carm offset is more successful - we get to slightly higher arm powers before the PRMI loses lock.  We should check if we think this will work in principle and we're just saturating something somewhere, or if 3f can't hold us to zero carm offset no matter what.

A note on technique:  We should be able to get the transfer function between MC2 actuation and ALS frequency by either a direct measurement, or Wiener filtering.  We need this in order to get the frequency subtraction to work in the correct units.

  11000   Wed Feb 11 03:41:12 2015 KojiUpdateLSCPRC error signal RF spectra

As the measurements have been done under feedback control, the lower RF peak height does not necessary mean
the lower optical gain although it may be the case this time.

These non-33MHz signals are embarassingly high!
We also need to check how these non-primary RF signals may cause spourious contributions in the error signals,
including the other PDs.

  10999   Wed Feb 11 02:42:05 2015 JenneUpdateLSCPRC error signal RF spectra

Since we're having trouble keeping the PRC locked as we reduce the CARM offset, and we saw that the POP22 power is significantly lower in the 25% MICH offset case than without a MICH offset, Rana suggested having a look at the RF spectra of the REFL33 photodiode, to see what's going on. 

The Agilent is hooked up to the RF monitor on the REFL33 demod board.  The REFL33 PD has a notch at 11MHz and another at 55MHz, and a peak at 33MHz. 

We took a set of spectra with MICH at 25% offset, and another set with MICH at 15% offset.  Each of these sets has 4 traces, each at a different CARM offset.  Out at high CARM offset, the arm power vs. CARM offset is pretty much independent of MICH offset, so the CARM offsets are roughly the same between the 2 MICH offset plots. 

What we see is that for MICH offset of 25%, the REFL33 signal is getting smaller with smaller CARM offset!!  This means, as Rana mentioned earlier this evening, that there's no way we can hold the PRC locked if we reduce the CARM offset any more. 

However, for the MICH offset 15% case, the REFL 33 signal is getting bigger, which indicates that we should be able to hold the PRC.  We are still losing PRC lock, but perhaps it's back to mundane things like actuator saturation, etc. 

The moral of the story is that the 3f locking seems to not be as good with large MICH offsets.  We need a quick Mist simulation to reproduce the plots below, to make sure this all jives with what we expect from simulation.

For the plots, the blue trace has the true frequency, and each successive trace is offset in frequency by a factor of 1MHz from the last, just so that it's easier to see the individual peak heights.

Here is the plot with MICH at 25% offset:

And here is the plot with MICH at 15% offset:

Note that the analyzer was in "spectrum" mode, so the peak heights are the true rms values.  These spectra are from the monitor point, which is 1/10th the value that is actually used.  So, these peak heights (mVrms level) times 10 is what we're sending into the mixer.  These are pretty reasonable levels, and it's likely that we aren't saturating things in the PD head with these levels. 

The peaks at 100MHz, 130MHz and 170MHz that do not change height with CARM offset or MICH offset, we assume are some electronics noise, and not a true optical signal.

Also, a note to Q, the new netgpib scripts didn't write data in a format that was back-compatible with the old netgpib stuff, so Rana reverted a bunch of things in that directory back to the most recent version that was working with his plotting scripts.  sorry.

 

Attachment 1: REFL33_25.pdf
REFL33_25.pdf
Attachment 2: REFL33_15.pdf
REFL33_15.pdf
  10998   Wed Feb 11 00:07:54 2015 ranaUpdateLSCLock Loss plot

Here is a lock loss from around 11 PM tonight. Might be due to poor PRC signals.  \oint {\frac{\partial PRCL}{\partial x}}

This is with arm powers of ~6-10. You can see that with such a large MICH offset, POP22 signal has gone done to zero. Perhaps this is why the optical gain for PRCL has also dropped by a factor of 30 crying.

This seems untenable no. We must try this whole thing with less MICH offset so that we can have a reasonable PRCL signal.cool

Attachment 1: 1107673198.png
1107673198.png
  10997   Tue Feb 10 18:37:17 2015 ericqUpdateASCQPD ASC ready to go

I've remeasured the QPD ASC sensing coefficients, and figured out what I did weird with the actuation coefficients. I've rearranged the controller filters to be able to turn on the boost in a triggered way, and written Up/Down scripts that I've tested numerous times, and Jenne has used as well; they are exposed on the ASC screen. 

All four loops (2 arms * pit,yaw), have their gains set for 8Hz UGF, and have 60 degrees of phase margin. The loop shape is the same as the previous ELOG post. Here is the current on/off performance. The PDH signals (not shown, but in attached xml) show no extra noise, and the low frequency RIN goes down a bit, whic is good. The oplevs error signals are a bit noisier, but I suppose that's unavoidable. The Y-arm performs a bit better than the X-arm. 

The up/down scripts don't touch the filters' trigger settings at all, just handles switching the input and output and clearing history. FM1 contains the boost which is intended to have a longer trigger delay than the filters themselves.

Attachment 1: Feb10_loops_offOn.png
Feb10_loops_offOn.png
Attachment 2: Feb10_newLoops_offOn.xml.zip
  10996   Tue Feb 10 16:01:21 2015 manasaUpdateGeneralAUX X fiber coupled 72%

Plan C finally worked. We have 1.454mW of AUX X light at the PSL table (2mW incident on the fiber coupler). 

Attached is the layout of the telescope.

What I did:

I stuck in Lens 1 (f=200mm) and measured the beam width after the focus of the lens at several points. I fit the data and calculated the beam waist and its position after this lens. 

I used the calculated waist and matched it with an appropriate lens and target (fiber coupler) distance. I calculated the maximum coupling efficiency to be 77% for Lens 2 with f=50mm and the fiber coupler placed at 20cm from the waist of Lens1. I was able to obtain 72% coupling after putting the telescope together.

I locked the arms, ran ASS and brought back GTRX to its usual optimum value of ~0.5 counts after closing. We also have the X arm beatnote on the spectrum analyzer.

Notes:

There are still a couple of things to fix. The rejected beam from the beam sampler has to be dumped using a razor blade.

 

  10995   Tue Feb 10 13:48:58 2015 manasaUpdateLSCProbable cause for headaches last night

I found the PSL enclosure open (about a feet wide) on the north side this morning. I am assuming that whoever did the X beatnote alignment last night forgot to close the door to the enclosure before locking attempts frown

Quote:

Unfortunately, we only had one good CARM offset reduction to powers of about 25, but then my QPD loop blew it. We spent the vast majority of the night dealing with headaches and annoyances. 

Things that were a pain:

  • If TRX is showing large excursions after finding resonance, there is no hope. These translate into large impulses while reducing the CARM offset, which the PRMI has no chance of handling. The first time aligning the green beat did not help this. For some reason, the second time did, though the beatnote amplitude wasn't increased noticibly. 
    • NOTICE: We should re-align the X green beatnote every night, after a solid ASS run, before any serious locking work. 
    • Afterwards, phase tracker UGFs (which depend on beatnote amplitude, and thereby frequency) should be frequently checked. 
  • We suffered some amount from ETMX wandering. Not only for realigning between lock attempts, but on one occasion, with CARM held off, GTRX wandered to half its nominal value, leading to a huge effective DARM offset, which made it impossible to lock MICH with any reasonble power in the arms. Other times, simply turning off POX/POY locking, after setting up the beatnotes, was enough to significantly change the alignment. 
  • IMC was mildly tempermental, at its worst refusing to lock for ~20min. One suspicion I have is that when the PMC PZT is nearing its rail, things go bad. The PZT voltage was above 200 when this was happening, after relocking the PMC to ~150, it seems ok. I thing I've also had this problem at PZT voltages of ~50. Something to look out for. 

Other stuff:

  • We are excited for the prospect of the FOL system, as chasing the FSS temperature around is no fun. 
  • UGF servo triggering greatly helps the PRMI reacquire if it briefly flashes out, since the multipliers don't run away. This exacerbated the ALS excursion problem. 
  • Using POPDC whitening made it very tough to hold the PRMI. Maybe because we didn't reset the dark offset...?

 

  10994   Tue Feb 10 03:09:02 2015 ericqUpdateLSCSome locking thoughts on PRMI

Unfortunately, we only had one good CARM offset reduction to powers of about 25, but then my QPD loop blew it. We spent the vast majority of the night dealing with headaches and annoyances. 

Things that were a pain:

  • If TRX is showing large excursions after finding resonance, there is no hope. These translate into large impulses while reducing the CARM offset, which the PRMI has no chance of handling. The first time aligning the green beat did not help this. For some reason, the second time did, though the beatnote amplitude wasn't increased noticibly. 
    • NOTICE: We should re-align the X green beatnote every night, after a solid ASS run, before any serious locking work. 
    • Afterwards, phase tracker UGFs (which depend on beatnote amplitude, and thereby frequency) should be frequently checked. 
  • We suffered some amount from ETMX wandering. Not only for realigning between lock attempts, but on one occasion, with CARM held off, GTRX wandered to half its nominal value, leading to a huge effective DARM offset, which made it impossible to lock MICH with any reasonble power in the arms. Other times, simply turning off POX/POY locking, after setting up the beatnotes, was enough to significantly change the alignment. 
  • IMC was mildly tempermental, at its worst refusing to lock for ~20min. One suspicion I have is that when the PMC PZT is nearing its rail, things go bad. The PZT voltage was above 200 when this was happening, after relocking the PMC to ~150, it seems ok. I thing I've also had this problem at PZT voltages of ~50. Something to look out for. 

Other stuff:

  • We are excited for the prospect of the FOL system, as chasing the FSS temperature around is no fun. 
  • UGF servo triggering greatly helps the PRMI reacquire if it briefly flashes out, since the multipliers don't run away. This exacerbated the ALS excursion problem. 
  • Using POPDC whitening made it very tough to hold the PRMI. Maybe because we didn't reset the dark offset...?
  10993   Tue Feb 10 02:55:29 2015 JenneConfigurationModern ControlQuacky filters

I discovered that somehow my Wiener filters that show up in Foton are not the same as what I have in Matlab-land. 

I have plotted each of my 3 filters that I'm working with right now (T-240 X, Y and Z for PRCL Pitch) at several stages in the filter creation process.  Each plot has:

  • Blue trace is the Wiener filter that I want, after the vectfit process.
  • Green trace is the frequency response of the SOS filters that are created by autoquack (really, quack_to_rule_them_all, which is called by autoquack).  The only thing that happens in matlab after this is formatting the coefficients into a string for writing to foton.
  • Red trace is the exported text file from foton.

It's not just a DC gain issue - there's a frequency dependent difference between these filters.  Why?? 

It's not frequency warping from changing between analog and digital filters.  The sample rate for the OAF model is 2048Hz, so the effect is small down at low frequencies.  Also, the green trace is already discretized, so if it were freq warping, we'd see it in the green as well as red, which clearly we don't.

Has anyone seen this before?  I'm emailing my seismic friends who deal in quack more than I do (BLantz and JKissel, in particular) to see if they have any thoughts.

Also, while I'm asking questions, can autoquack clear filters?  Right now I'm overwriting old filters with zpk([],[],1)'s, which isn't quite the same.  (I need this because the Wiener code needs more than one filter module to fit all of the poles I need, and it decides for itself how many FMs it needs by comparing the length of the poles vector with 20.  If one iteration needs 4 filter modules, but the next iteration only wants 3, I don't want to leave in a bogus 4th filter. 

Here are the plots:

(The giant peak at ~35Hz in the Z-axis fiilter is what tipped me off that something funny was going on)

  10992   Tue Feb 10 02:40:54 2015 JenneUpdateLSCSome locking thoughts on PRMI

[EricQ, Jenne]

We wanted to make the PRMI lock more stable tonight, which would hopefully allow us to hold lock much longer.  Some success, but nothing out-of-this-world.

We realized late last week that we haven't been using the whitening for the ASDC and POPDC signals, which are combined to make the MICH error signal.  ASDC whitening is on, and seems great.  POPDC whitening (even if turned on after lock is acquired) seems to make the PRMI lock more fussy.  I need to look at this tomorrow, to see if we're saturating anything when the whitening is engaged for POPDC.

One of the annoying things about losing the PRMI lock (when CARM and DARM have kept ALS lock) is that the UGF servos wander off, so you can't just reacquire the lock.  I have added triggering to the UGF servo input, so that if the cavity is unlocked (really, untriggered), the UGF servo input gets a zero, and so isn't integrating up to infinity.  It might need a brief "wait" in there, since any flashes allow signal through, and those can add up over time if it takes a while for the PRMI to relock.  UGF screens reflect this new change.

  10991   Mon Feb 9 17:47:17 2015 diegoUpdateLSCCM servo & AO path status

I wrote the script with the recipe we used, using the Yarm and AS55 on the IN2 of the CM board; however, the steps where the offset should be reduced are not completely deterministic, as we saw that the initial offset (and, therefore, the following ones) could change because of different states we were in. In the script I tried to "servo" the offset using C1:LSC-POY11_I_MON as the reference, but in the comments I wrote the actual values we used during our best test; the main points of the recipe are:

  • misalign the Xarm and the recycling mirrors;
  • setting up CARM_B for POY11 locking and enabling it;
  • setting up CARM_A for CM_SLOW;
  • setting up the CM_SLOW filter bank, with only FM1 and FM4 enabled;
  • setting up the CARM filter bank: FM1 FM2 FM6 triggered, only FM3 and FM5 on; usual CARM gain = 0.006;
  • setting up CARM actuating on MC2;
  • turn off the violin filter FM6 for MC2;
  • setting up the default configuration for the Common Mode Servo and the Mode Cleaner Servo; along with all the initial parameters, here is where the initial offset is set;
  • turn on the CARM output and, then, enable LSC mode;
  • wait until usual POY11 lock is acquired and, a bit later, transition from CARM_B to CARM_A;
  • then, the actual CM_SLOW recipe:
    • CM_AO_GAIN = 6 dB;
    • SUS-MC2_LSC FM6 on (the 300:80 filter);
    • CM_REFL2_GAIN = 18 dB;
    • servo CM_REFL_OFFSET;
    • CM_AO_GAIN = 9 dB;
    • CM_REFL2_GAIN = 21 dB;
    • servo CM_REFL_OFFSET;
    • CM_REFL2_GAIN = 24 dB;
    • servo CM_REFL_OFFSET;
    • CM_REFL2_GAIN = 27 dB;
    • servo CM_REFL_OFFSET;
    • CM_REFL2_GAIN = 31 dB;
    • servo CM_REFL_OFFSET;
    • CM_AO_GAIN = 6 dB;
    • SUS-MC2_LSC FM7 on (the :300 compensating filter);

I tried the procedure and it seems fine, as it did during the tries Q and I made; however, since it touches many things in many places, one should be careful about which state the IFO is into, before trying it.

The script is in scripts/CM/CM_Servo_OneArm_CARM_ON.py and in the SVN.

 

  10990   Mon Feb 9 17:23:17 2015 diegoUpdateComputer Scripts / ProgramsNew laptops

I forgot to elog about these ones, my bad... The new/updated laptops are giada, viviana and paola; paola is already in the lab, while giada and viviana are in the control room waiting for a new home. The Pool of Names Wiki page has already been updated to reflect the changes.

  10989   Mon Feb 9 08:40:49 2015 SteveUpdateSUSrecent earthquake 4.9

Baja 4.9 m earth quake tripped suspentions, except ETMX Sus damping recovered. MC is locking.
 

Attachment 1: M4.9Baja.png
M4.9Baja.png
  10988   Sun Feb 8 21:54:50 2015 ranaSummaryPSLISS AOM driver check

This is good news. It means that the driver probably won't limit the response of the loop - I expect we'll get 20-30 deg of phase lag @ 100 kHz just because of the acoustic response of the AOM PZT + crystal.

  10987   Sat Feb 7 21:30:45 2015 JenneUpdateLSCPRC aligned

I'm leaving the PRC aligned and locked.  Feel free to unlock it, or do whatever with the IFO.

  10986   Sat Feb 7 13:34:11 2015 KojiSummaryPSLISS AOM driver check

I wanted to check the status of the ISS. The AOM driver response was measured on Friday night.
The beam path has not been disturbed yet.

- I found the AOM crystal was removed from the beam path. It was left so.

- The AOM crystal has +24V power supply in stead of specified +28V.
  I wanted to check the functionality of the AOM driver.

- I've inserted a 20dB directional coupler between the driver and the crystal.
  To do so, I first turned off the power supply by removing the corresponding fuse block at the side panel of the 1X1 Rack.
  Then ZFDC-20-5-S+ was inserted, the coupled output was connected to a 100MHz oscilloscope with 50Ohm termination.
  Then plugged in the fuse block again to energize the driver box.

  Note that the oscilloscope bandwidth caused reduction the amplitude by a factor of 0.78. In the result, this has already been compensated.

- First, I checked the applied offset from a signal generator (SG) and the actual voltage at the AOM input. The SG OUT
  and the AOM control input are supposed to have an impedance of 50Ohm. However, apparently the voltage seen at the
  AOM in was low. It behaved as if the input impedance of the AOM driver is 25Ohm.
  In any case, we want to use low output impedance source to drive the AOM driver, but we should keep this in mind.

- The first attachment shows the output RF amplitude as a function of the DC offset. The horizontal axis is the DC voltage AT THE AOM INPUT (not at the SG out).
  Above 0.5V offset some non linearity is seen. I wasn't sure if this is related to the lower supply voltage or not. I'd use the nominal DC of 0.5V@AOM.

  The output with the input of 1V does not reach the specified output of 2W (33dBm). I didn't touch the RF output adjustment yet. And again the suppy is not +28V but +24V.

- I decided to measure the frequency response at the offset of 0.53V@AOM, this corresponds to the DC offset of 0.8V. 0.3Vpp oscillation was given.
  i.e. The SG out seen by a high-Z scope is V_SG(t) = 1.59 + 0.3 Sin(2 pi f t) [V]. The AOM drive voltage V_AOM(t) = 0.53 + 0.099 Sin(2 pi f t).
  From the max and min amplitudes observed in the osciiloscope, the response was checked. (Attachment 2)
  The plot shows how much is the modulation depth (0~1) when the amplitude of 1Vpk is applied at the AOM input.
  The value is ~2 [1/V] at DC. This makes sense as the control amplitude is 0.5, the applied voltage swings from 0V-1V and yields 100% modulation.

  At 10MHz the first sign of reduction is seen, then the response starts dropping above 10MHz. The specification says the rise time of the driver is 12nsec.
  If the system has a single pole, there is a relationship between the rise time (t_rise) and the cut-off freq (fc) as fc*t_rise = 0.35 (cf Wikipedia "Rise Time").
  If we beieve this, the specification of fc is 30MHz. That sounds too high compared to the measurement (fc ~15MHz).
  In any case the response is pretty flat up to 3MHz.

Attachment 1: AOM_drive.pdf
AOM_drive.pdf
Attachment 2: AOM_response.pdf
AOM_response.pdf
  10985   Fri Feb 6 18:06:18 2015 manasaUpdateGeneralFiber Optic module for FOL

I pulled out the Fiber Optic Module for FOL from the rack inside the PSL table enclosure and modified it. The beat PDs were moved into the box to avoid breaking the fiber pigtail input to the PD.

The box has 3 input FC/APC connectors (PSL and AUX lasers) and 2 output FC/APC connectors (10% of the beatnote for the AUX lasers).

Attachment shows what is inside the box. The box will again go back on the rack inside the PSL enclosure.

Attachment 1: FOLfiberModule.png
FOLfiberModule.png
  10984   Fri Feb 6 15:29:29 2015 ericqUpdateCDSNetwork Shenanigans

Just now we had another EPICS freeze. The network was still up; i.e. I could ssh to chiara and fb, who both seemed to be working fine.

I could ping c1lsc successfully, but ssh just hung. fb's dmesg had some daqd segfault messages, so I telnet'ed to daqd and shut it down. Soon after, EPICS came back, but this is not neccesarily because of the daqd restart...

  10983   Fri Feb 6 10:03:23 2015 SteveUpdatePEMdusty days

 

Quote:

Yesterday morning was dusty. I wonder why?

The PRM sus damping was restored this morning.

Yesterday afternoon at 4 the dust count peaked 70,000 counts

Manasa's alergy was bad at the X-end yesterday. What is going on?

There was no wind and CES neighbors did not do anything.

Attachment 1: duststorm.png
duststorm.png
  10982   Fri Feb 6 03:21:17 2015 diegoUpdateLSCCARM Transition to REFL11 using CM_SLOW Path

[Diego, Jenne]

We kept struggling with the PRMI, although it was a little better than yesterday:

  • we retuned the X Green beatnote;
  • we managed to reach lower CARM offsets than yesterday night, but we still can't keep lock long enough to perform a smooth transition to CM SLOW/REFL11;
  • we tweaked MICH a bit:
    • the ELP in FM8 now is always on, because it seems to help;
    • we tried using a new FM1 1,1:0,0 instead of FM2 1:0 because we felt we needed a little more gain at low frequencies, but unfortunately this didn't change much MICH's behaviour;
    • now, after catching PRMI lock, the MICH limiter is raised to 30k (in the script), as a possible solution for the railing problem; the down/relock scripts take care of resetting it to 10k while not locked/locking;

So, still no exciting news, but PRMI lock seems to be improving a little.

  10981   Thu Feb 5 15:21:25 2015 manasaUpdateGeneralX endtable work

[EricG, Manasa]

We were at the X end today trying to couple AUX X light into the fiber. 

The proposed plan still did not give a good beampath. The last steering mirror before the fiber coupler was sticking out of the table enclosure. I tried a few other options and the maximum coupling that I could get was ~10%

I am working on plan C now; which would be to use fixed mount mirrors and steer the beam to the space created by Koji near the IR trans path and use a set of lenses instead of a single lens. I will elog more details after some modematching calculations.

We moved one of the clamps for the doubling crystal to make space. Also, the NPRO current was reduced during this work.

I reset things to how they were before I touched the table. I ensured that the green power was still the same (~3mW) after the doubler and that it could lock to the arm in TEM00.

  10979   Thu Feb 5 04:35:14 2015 diegoUpdateLSCCARM Transition to REFL11 using CM_SLOW Path

[Diego, Eric]

Tonight was a sad night... We continued to pursue our strategy, but with very poor results:

  • before doing anything, we made sure we had a good initial configuration: we renormalized the arm powers, retuned the X arm green beatnote, did extensive ASS alignment;
  • since the beginning of the night we faced a very uncooperative PRMI, which caused a huge number of locklosses, often just by itself, without even managing to reduce the MICH offset before reducing the CARM one;
  • we had to reduce the PRCL gain to -0.002 in order to acquire PRMI lock, but keeping it such or restoring it to -0.004 once lock was acquired either didn't improve the PRMI stability at all;
  • we also tweaked a bit the PRCL and MICH UGF servos (namely, their frequencies to ~80 Hz and ~40 Hz respectively) and that seemed to help earlier during the night, but not much longer;
  • we only managed to transition CARM to REFL11 via CM SLOW twice;
    • the first time we lost lock almost immediately, probably because of a non-optimal offset between CARM A and B;
    • the second time we managed to stay there a little longer, but then some spike in the PRCL loop and/or the MICH loop hitting the rails threw us out of lock (see the lockloss plot);
    • both times we transitioned at arm power ~18;
  • during the night we used an increased analog ASDC whitening gain, as from Eric's elog here http://nodus.ligo.caltech.edu:8080/40m/10972 ; even with this fix, though, MICH is still often hitting the rails and causing the lock losses;
  • the conclusion for tonight is that we need to figure what is going on with the PRMI...

 

Attachment 1: 4Feb2015_Transition_CARM_REFL11_CM_SLOW_AP_18.png
4Feb2015_Transition_CARM_REFL11_CM_SLOW_AP_18.png
  10978   Wed Feb 4 21:09:43 2015 manasaUpdateGeneralX endtable work scheduled tomorrow

The X end fiber setup will be put together tomorrow morning. Let me know if there are any concerns.

Quote:

It is certain that we have space issues at the X end that has been preventing us from sticking in a lens to couple light into the fiber. 

The only way out is to install a platform on the table where we can mount the lens. I have attached the a photo of how things look like at the X end (attachment 1) and also a drawing of the platform that which can hold the lens (attachment 2). Additional support to the raised platform will be added depending on how much space we can clear up on the table by moving the clamping forks of the doubler.
Steve and I have been able to gather parts that can be put together into something similar to what is shown in the drawing. 

Proposed modifications to the X end table:

1. The side panels of the table enclosure will come out while putting in the new platform.

2. The clamping forks for the doubling crystal will be moved.

Let me know of any concerns about the proposed solution. 

 

  10977   Wed Feb 4 21:05:24 2015 manasaUpdateGeneralRF amplifier for ALS

The components of the RF amplifier box are in place. The RF amplifier box has been mounted on the IOO rack and the front panel connections have been labeled. Attached is the photo of how things look in the inside for future reference.

Sometime in the next few days the box will be pulled out to replace the panel mount SMA barrels in the front with insulated ones.

Attachment 1: RFampBox.png
RFampBox.png