I tried many times this evening to engage the AO path, with limited success.
Q's new scripts worked really well, and so I have some transfer functions! To take these measurements, in ...../scripts/general/netgpib, I am running ./TFSR785 TFSR785_CARMloop_May2014.yml , where the file name is the name of my parameter file. The data, and the saved pdfs, are in /users/jenne/PRFPMI/CARM_loop_measurements/2014May14/ . For these measurements, the SR785 is hooked up to the "A" set of excitation and test points on the CM board.
All of these traces were taken while the IFO was PRFPMI, with PRCL and MICH on REFL33, DARM on ALS diff, and CARM on InvSqrtTrans. carm_cm_up.sh is up to date, through the echo "REFL_I should now be zero" (~line 111 in the script). All you need to do is set the beatnotes, and then run the script. Follow instructions in the prompt (such as "press enter to confirm PRMI is locked").

Here are my notes for the various times:
23:01:44 - MC IN2 = 0dB, CARM gain = 5.0
23:13:45 - MC IN2 = 10dB, CARM gain = 5
23:26:10 - MC IN2 = 6dB, CARM gain = 10 (after Q suggested increasing overall gain, rather than just AO path)
00:13:07 - MC IN2 = 6dB?, CARM = 6ish? don't remember exactly.
00:45:00ish, Realigned IFO using IR with arms.
01:03:17 - MC In2 = 0dB, CARM gain = 5
01:07:42 - MC IN2 = 8dB, CARM gain = 6.295 (AO went up to 6dB, then +1dB steps to both simultaneously using ezcastep C1:LSC-CARM_GAIN 1dB C1:IOO-MC_AO_GAIN 1 )
01:08:57 - MC IN2 = 10dB, CARM gain = 7.92447
01:10:08 - MC IN2 = 12dB, CARM gain = 9.97631
lockloss when trying to add 1 more dB to both.
01:41:36 - MC IN2 = 12dB, CARM gain = 9.97
lockloss when just MC IN2 up by 1dB, left CARM gain alone.
Other notes:
The 60Hz noise in TRY is back. Since I thought I remembered someone suggesting that it was leakage light from the exit sign, Koji went in and wrapped the end table in foil, however the lines are still present. |