40m
QIL
Cryo_Lab
CTN
SUS_Lab
CAML
OMC_Lab
CRIME_Lab
FEA
ENG_Labs
OptContFac
Mariner
WBEEShop
|
40m Log |
Not logged in |
 |
|
Tue May 24 11:04:53 2022, JC, Update, VAC, Beginning Pumpdown 
|
Wed May 25 16:38:27 2022, yuta, Update, SUS, Suspensions quick health check 
|
Tue Jun 7 17:26:21 2022, yuta, Update, SUS, ITMY ULCOIL mystery not solved  
|
Tue Jun 7 19:13:12 2022, yuta, Update, SUS, Agreement on suspension damping loop polarity conventions
|
Fri Jun 10 15:02:04 2022, yuta, Update, SUS, Servo gain sign flipped for MC1 WFS relief
|
Tue Jun 7 19:40:45 2022, Anchal, Update, SUS, ITMY changed output matrix to disable use of UL coil 
|
Wed Jun 8 18:16:20 2022, yuta, Update, SUS, ITMY ULCOIL mystery: Coil driver swap test 
|
Thu Jun 9 23:08:39 2022, rana, Update, SUS, ITMY ULCOIL mystery: Coil driver swap test
|
Fri Jun 10 13:02:14 2022, yuta, Update, SUS, ITMY ULCOIL mystery: Coil driver swap test
|
Fri Jun 10 15:04:23 2022, rana, Update, SUS, ITMY ULCOIL mystery: Coil driver swap test
|
Fri Jun 10 21:10:01 2022, yuta, Update, SUS, ITMY ULCOIL mystery: Coil driver swap test
|
Wed Jun 15 17:03:17 2022, yuta, Update, SUS, ITMY ULCOIL issue solved, loose connection in sat amp box   
|
Wed Jun 15 15:45:37 2022, yuta, Update, SUS, LO1 LLCOIL now working, it was loose connection  
|
Wed Jun 15 07:26:35 2022, JC, Update, VAC, Beginning Pumpdown
|
|
Message ID: 16920
Entry time: Wed Jun 15 17:03:17 2022
In reply to: 16910
|
Author: |
yuta |
Type: |
Update |
Category: |
SUS |
Subject: |
ITMY ULCOIL issue solved, loose connection in sat amp box |
|
|
[Anchal, Yuta]
We fixed the issue of ITMY ULCOIL not driving ITMY by replacing one of the 64pin ribbon cable in the satellite amplifier box.
We thought the coil driver and the sat amp box are OK by checking the voltage change at the output of the sat amp box by giving an offset to UL coil driver, but it was not giving a current change, probably due to too much contact resistance in the cables.
It was sneaky because it was not completely disconnected.
All the coils for our suspensions are now working!
What we did:
- Using breakout boards, the output current of sat amp box was measured using FLUKE multimeter. It turned out that UL is not giving measurable current. We also confirmed that UR coil driver can drive UL by re-directing the current from UR coil driver to UL. This means that the UL magnet was not de-magnetized!
- Measured the coil resistance from at the coil driver output and found that UL coil seen from there has too high resistance which cannot be measured with the multimeter, whereas UR coil was measured to be ~30 Ohms.
- Went back to the feedthru and measured the resistance of UL coil. Upto the output of the Satellite Amp Terimator, the resistance was measured to be ~16 Ohms, but not at the input of the Satellite Amp Terimator (Attachment #1,2).
- It turned out that #16 pin of 64pin ribbon cable in between the Satellite Amp Terimator (LIGO-D990021) and the Satellite Amp board (LIGO-D961289) at the Satellite Amp Terimator side was not good (Attachment #3).
- Replaced the cable and confirmed that ULCOIL can kick ITMY (Attachment #4).
- C1:SUS-ITMY_TO_COIL matrix was reverted to default values.
Next:
- We might have to re-commission Yarm ASS again since pitch-yaw coupling have changed. -> EDIT: Checked that it works (except for ITM PIT L), including offloading offsets (writeASS_offsets.py), 18:30 local.
- Now that LO1 LLCOIL issue is solved and LO2 stuck is solved, we should do the free swing test again to identify the resonant frequencies.
- OSEM sensor diagonalization (input matrix), coil balancing (and F2A) |
|
|
|
|