I did all these checks today. 

Quote:

I will check (i) REFL55 transimpedance, (ii) cable loss between AP table and 1Y2 and (iii) is the beam well centered on the REFL55 photodiode.

  1. The transimpedance was measured to be ~420 ohms at 55 MHz (-4.3 dB relative to the assumed 700V/A of the NF1611), so close to what I measured in June (the data download didn't work apparently and so I don't have a plot but it can readily be repeated). The DC levels also checked out - with 20mA drive current for the Jenne laser, I measured ~2.3 V on the NF1611 (10kohm DC transimpedance) vs ~13mV on the DC output of the REFL55 PD (50 ohm DC transimpedance).
  2. Time domain confirmation of the above statement is seen in Attachment #1. The Agilent was used to drive the Jenne laser with 0dBm RF signal @ 55 MHz. Ch1 (yellow) is the REFL55 PD output, Ch2 (blue) is the NF1611 RFPD, measured at the AP table (sorry for the confusing V/div setting).
  3. Re-connected the cabling at the AP table, and measured the signal at 1Y2 using the scope Rana conveniently left there, see Attachment #2. Though the two scopes are different, the cable+connector loss estimated from the Vpp of the signal at the AP table vs that at 1Y2 is 1.5 dB, which isn't outrageous I think.
  4. For the integrated test, I left the AM laser incident on the REFL55 photodiode, reconnected all the cabling to the CDS system, and viewed the traces on ndscope, see Attachment #3. Again, I think all the numbers are consistent. 
    • REFL55 demod board has an overall conversion gain (including the x10 gain of the daughter board preamp) of ~5V I/F per 1V RF.
    • There is a flat 18 dB whitening gain.
    • The digitized signal was ~13000 ctspp - assuming 3276.8 cts/V, that's ~4Vpp. Undoing the flat whitening gain and the conversion efficiency, I get 13000 / 3276.8 / (10^(18/20)) / 5 ~ 100 mVpp, which is in good agreement with Attachment #3 (pardon the thin traces, I didn't realize it looked so bad until I closed everything).

So it would seem that there is nothing wrong with the sensing electronics. I also think we can rule out any funkiness with the modulation depths since they have been confirmed with multiple different measurements.

One thing I checked was the splitting ratios on the AP table. Jenne's diagram is still accurate (assuming the components are labelled correctly). Let's assume 0.8 W makes it through the IMC to the PRM - then, I would expect, according to the linked diagram, 0.8 W * 0.8 * (1-5.637e-2) * 0.8 * 0.1 * 0.5 * 0.9 ~ 22 mW to make it onto the REFL55 PD. With the PRM aligned and the beam centered on the PD (using DC monitor but I also looked through an IR viewer, looked pretty well centered), I measured 500 mV DC level. Assuming 50 ohm DC transimpedance, that's 500 / 50 / 0.8 ~ 12.5 mW of power on this photodiode, which while is consistent with what's annotated on Jenne's diagram, is ~50% off from expectation. Is the uncertainty in the Faraday transmission and IMC transmission enough to account for this large deviation?

If we want more optical gain, we'd have to put more light on this PD. I suppose we could have ~10x the power since that's what is on IMC REFL when the MC is unlocked? If we want x100 increase in optical gain, we'd also have to increase the transimpedance by 10. I'll double check the simulation but I"m inclined to believe that the sensing electronics are not to blame.


Unconnected to this work but I feel like I'm flying blind without the wall StripTool traces so I restored them on zita (ran /opt/rtcds/caltech/c1/scripts/general/startStrip.sh).

Attachment 1: IMG_9140.jpg  3.163 MB  Uploaded Fri Mar 19 23:30:04 2021  | Hide | Hide all
IMG_9140.jpg
Attachment 2: IMG_9141.jpg  3.142 MB  Uploaded Fri Mar 19 23:30:16 2021  | Hide | Hide all
IMG_9141.jpg
Attachment 3: REFL55.png  202 kB  Uploaded Fri Mar 19 23:34:28 2021  | Hide | Hide all
REFL55.png
ELOG V3.1.3-