Entry  Mon Mar 27 20:41:21 2017, rana, gautam, Update, PSL, PMC DAQ assay for feed-forward integration TEK00000.PNG20170327_194931.jpg20170327_204554.jpg
    Reply  Wed Mar 29 11:41:19 2017, gautam, Update, PSL, PMC DAQ assay for feed-forward integration C1ALS_PSLnamespace.pngPMCcheckout.pdfD980352-A-40m_151119.pdf
       Reply  Thu Mar 30 00:16:09 2017, gautam, Update, PSL, PSL NPRO PZT calibration 
          Reply  Mon Apr 3 17:25:13 2017, gautam, Update, PSL, PSL NPRO PZT calibration PDHerr.pdfNPROcalib.pdf
             Reply  Tue Apr 4 17:27:58 2017, rana, Update, PSL, PSL NPRO PZT calibration 
       Reply  Thu Mar 30 18:11:01 2017, rana, Update, PSL, PMC DAQ assay for feed-forward integration 
          Reply  Mon Apr 3 23:07:09 2017, gautam, Update, PSL, PMC DAQ assay for feed-forward integration 
             Reply  Tue Apr 11 00:38:37 2017, gautam, Update, PSL, PMC demod moved off servo board PMC_Ctrl_spec.pdf
                Reply  Wed Apr 12 00:36:53 2017, gautam, Update, PSL, PMC demod moved off servo board PMC_digitalSpec.pdfPMC_DAQ_spectra.pdf
                   Reply  Tue Apr 18 01:01:03 2017, gautam, Update, PSL, PMC OLTF measured, DAQ channels calibrated PMCspectra_calibrated.pdfPMC_MEDMs.png
                      Reply  Tue Apr 18 16:10:00 2017, gautam, Update, PSL, PMC OLTF measured, DAQ channels calibrated PMC_OLTF_170418.pdf
                         Reply  Tue Apr 18 23:37:15 2017, rana, Update, PSL, PMC OLTF measured, DAQ channels calibrated 
                         Reply  Wed Apr 19 15:13:30 2017, gautam, Update, PSL, PMC/MCL multicoherence PMC_MCL_multicoherence.pdf
Message ID: 12911     Entry time: Mon Mar 27 20:41:21 2017     Reply to this: 12916
Author: rana, gautam 
Type: Update 
Category: PSL 
Subject: PMC DAQ assay for feed-forward integration 

We are thinking to use the PMC signals to help us in figuring out the feedback / feedforward stuff and making it better.

Today we scoped out the PMC DAQ channels (which were never re-hooked up after the Joe/Jamie CDS upgrade 6 years ago).

There is a 4-pin LEMO connector on the front panel which gives

  1. the error signal (after the 4th order, post-mixer lowpass and a OP27 buffer with a 17 kHz low pass)
  2. the feedback voltage to the PZT, after a resistive divide by 50

Both of these signals are buffered by the AD620 inst amp configured with a gain of 1. In the green scope trace, you can see that there's a ~110 MHz signal strongly evident there. In the spectrum analyzer screen shot there is a instrument noise trace and then a PMC error point trace. You can see that all the peaks are ony there when I connect to the servo board instead of a Terminator. This RF noise is mainly the higher harmonics of the 35.5 MHz modulation getting there. It seems to be in both the error and control DAQ outputs, and a question is whether or not it is also in the servo electronics.

I also attach a close up of the servo board in the region of the post-mixer LC low pass filtering. I think its supposed to be 4th order cutoff at 1 MHz, but maybe the caps are busted or there's a way for the RF from the mixer to bypass the filters and get into the main servo path?

In the medium term, we probably want to use the new PDH servo that Rich is making. Need to buy/make a HV driver to use, but that should be easy.