Used BeamR and WinCamD to profile the two light sources (ECDL and OPO pump)
(1) ECDL; profile 19** nm beam after the aspheric lens. I guess we want this beam to be nominally collimated for optimal feedback with the Littrow-configured grating, so I recorded the 1/e^2 waists (x, y) as a function of longitudinal displacement. The result is attached below. Linear fits provide rough estimates for the beam divergences, giving 2.0 mrad along x (parallel to the table) and 1.2 mrad along y (normal to the table) suggesting some astigmatism which is common in high NA aspheric lenses. I inspected the distance from the aspheric lens to the SAF gain chip and measured ~ 2.0 mm (compared to the 1.99 mm working distance specified for this lens with NA=0.71). The SAF1900S specifies a beam divergence angle of 35 deg (corresponding to NA=0.57), so there is room for improvement by tweaking the aspheric flexure alignment.
(2) NPRO; profile 1064 nm beam at low power (~10 mW) right after the head output. Having 10x more power made things way easier for this as compared to the ECDL, but the method was the same (record 1/e^2 waists as a function of longitudinal displacement). The result is attached below. Linear fits provide rough estimates for the beam divergences, giving 2.1 mrad along x, and 2.1 mrad along y. Here I grabbed the specified divergence of 2.3 mrad from a relatively old manual, and even drew the displaced waist profile (w0 = 160 um) which seemingly fit the profile, but the actual values may be different. |