  2872   Mon May 3 16:53:27 2010 josephbUpdateCDSUpdated lsc.mdl and the ifo plant model with memory locations

I've updated the LSC and IFO models that Rana created with new shared memory locations.  I've used the C1:IFO- for the ifo.mdl file outputs, which in turn are read by the lsc.mdl file.  The LSC outputs being lsc control signals are using C1:LSC-.  Optics positions would presumably be coming from the associated suspension model, and am currently using SUP, SPX, and SPY for the suspension plant models (suspension vertex, suspension x end, suspension y end).

I've updated the web view of these models on nodus.  They can be viewed at: https://nodus.ligo.caltech.edu:30889/FE/

I've also created a C1.ipc file in /cvs/cds/caltech/chans/ipc  which assigns ipcNum to each of these new channels in shared memory.

  11745   Tue Nov 10 02:34:28 2015 gautamUpdateLSCUpdated interpretation of peaks

After thinking about the interpretation of the various peaks seen in the scan through 2 FSRs, I have revised the information presented in the previous elog. Yutaro pointed out that the modulation frequency isn't exactly 11MHz, but according to this elog, is 11.066209 MHz. So instead of using mod(11e6,FSR), I really should have been using mod(11.066209,FSR) and mod(5*11.066209,FSR) to locate the positions of the 11MHz and 55MHz sidebands relative to the carrier resonances. With this correction, the 'unknown' peaks identified in Attachment #1 in elog 11743 are in fact the 55MHz sideband resonances. 

However, this means that the peaks which were previously identified as 55MHz sideband resonances have to be interpreted now - I'm having trouble identifying these. If we assume that the types of peaks present in the scan are 11 MHz sideband, 55MHz sideband, and the TEM00, TEM10, TEM20, TEM30, and TEM40 mode resonances, then the peaks marked in grey in Attachment #1 to this elog can be interpreted as TEM30 (right of a carrier resonance) and TEM40 (left of a carrier resonance) mode resonances - however, the fitted center frequencies differ from the expected center frequencies (determined using the same method as elog 469) by ~3% (for TEM30) and ~20% (for TEM40) - therefore I am skeptical about these peaks, particularly the 4th HOM resonances. In any case, they are the smallest of all the peaks, and any correction due to them will be small. 

The updated modulation depths are as follows (computed using the same method as described in elog 11743, the updated plot showing the ratio of bessel functions as a function of the modulation depth is Attachment #2 in this elog):

@11.066209 MHz ---- 0.179

@5*11.066209 MHz --- 0.226

These numbers are now reasonably consistent with those reported in elog10211.

As for the mode-matching efficiency, the overall number is almost unchanged if I assume the TEM30 peaks are accurately interpreted: 92.11%. But the dominant HOM contribution comes from the first HOM resonance: (TEM00 = 1, TEM20 = 0.0325, TEM10 = 0.0475, TEM30 =  0.0056). These numbers may change slightly if the 4th HOM resonances are also correctly identified.

ETMx is still not well behaved and the mode cleaner isnt too happy either, so I think we will save the measurement of the round trip arm loss for daytime tomorrow.  

  11746   Tue Nov 10 11:06:02 2015 yutaroUpdateLSCUpdated interpretation of peaks
Quote:
Quote:

- modulation depth = 0.390 +/- 0.062

There are two modulation frequencies that make it to the arm cavities, at ~11MHz and ~55MHz. Each of these will have their own modulation depth indepedent of each other. Bundling them together into one number doesn't tell us what's really going on. 

 

I'm sorry. I misunderstood two things when writing elog 11741: the number of modulation frequencies, and how to distinguish modulation peaks and HOM peaks.

 

Now, I report about interpretation of the peaks marked in grey in Attachment #1 in elog 11745.

Summary:

The peaks marked in grey are interpreted as 3rd and 4th HOM resonance, if we assume that the radius of curvature of ETMY is slightly different from measured value. (measured: 57.6 m --> assumed: 59.3 m)

 

What I have done:

I plotted the differences in frequency between HOM peaks and 00 mode peaks (see Attachment #1) vs. expected orders of modes. The plot is shown in Attachement #2.

By fitting these data points, I calculated most likely value of gradient of this plot. This value corresponds: g_ITMYg_ETMY=0.3781. However, measured data of the radii of curvature suggests that g_ITMYg_ETMY=0.358. Assuming that this disagreement comes from the difference between measured and real values of ROC of ETMY (ITM is almost flat so that change of ROC of ITM doesn't have significant effect on g_ITMg_ETM), ROC of ETMY should be 59.3 m, different from measured value 57.6 m.

 

What I'd like to know:

-- Is such disagreement of ROC usual? Could it happen?

-- Are there any other possible explanations for this disagreement (or interpretations of the peaks marked in grey)?

  11747   Tue Nov 10 11:40:03 2015 KojiUpdateLSCUpdated interpretation of peaks

What is the uncertainty of your RoC estimation?

One measurement of the ETMY ROC was 57.6m, but we trust another measured value of 60.26m than the other.
The value is always dependent on the spotposition on the mirror and how the ROC is calculated from the mirror phase map (e.g. spotsize, averaging method).
So I don't think this is a huge deviation from the spec.

  11748   Tue Nov 10 11:41:56 2015 KojiUpdateLSCUpdated interpretation of peaks

FYI: I've also reported the similar mod depths of

11M: 0.194
55M: 0.234

in ELOG11036 with a different kind of measurement method.

  11749   Tue Nov 10 16:34:00 2015 yutaroUpdateLSCUpdated interpretation of peaks
Quote:

What is the uncertainty of your RoC estimation?

The uncertainty came from the residual of linear fitting and based on my estimation,

ROC_ETMY = 59.3 +/- 0.1 m.

 

And I attach the updated figure of the fitting (with residual zoomed up).

Data points in the lower are (intentionally) slightly shifted horizontally to make it easy for us to see them.

It is hard, I think, to estimate the error of each data point, but I used 17 kHz for the errors of all data points; 17 kHz is the error of FSR estimation of this measurement, and since FSR is the distance between two carrier peaks we can consider that HOM distances, which are the distance between carrier peaks and HOM peaks, have similar order errors comared with that of FSR. 

  4212   Thu Jan 27 15:16:43 2011 josephbUpdateCDSUpdated generate_master_screens.py

I modified the generate_master_screens.py script in /opt/rtcds/caltech/c1/medm/master/ to handle changing the MCL (and MC_L) listings to ALS for the two ETM suspension screens and associated sub-screens.

The relevant added code is:

custom_optic_channels = ['ETMX',
{'MCL':'ALS','MC_L':'ALS'},
'ETMY',
{'MCL':'ALS','MC_L':'ALS'}]

 

for index in range(len(custom_optic_channels)/2):
   if optic == custom_optic_channels[index*2]:
     for swap in custom_optic_channels[index*2+1]:
       sed_command = start_sed_string + swap + "/" + custom_optic_channels[index*2+1][swap] + middle_sed_string + optic + file
       os.system(sed_command)

When run, it generates the correctly named C1:SUS-ETMX_ALS channels, and replaces MCL and MC_L with ALS in the matrix screens.

 

  10383   Thu Aug 14 14:58:03 2014 JenneUpdateGeneralUpdated game plan

2014_Aug_14.pdf

(Updated as of 4pm)

  10385   Thu Aug 14 15:42:29 2014 KojiUpdateGeneralUpdated game plan

 - ALS

End PDH UGF improvement / post mixer LPF investigation (with in 2 weeks)

- MC/FSS

Riju measured the MC REFL PD transimpedance. See ELOG and related.

- ASC

Why do we want to see less PRM motion? I thought PRC motion was causing
LSC issue of the central part. We wanted to maximize the PRM effect, don't we?
(Or is this to supress ETM motion during full lock?)

  10386   Thu Aug 14 15:51:37 2014 JenneUpdateGeneralUpdated game plan

Quote:

 - ALS

End PDH UGF improvement / post mixer LPF investigation (with in 2 weeks)

- MC/FSS

Riju measured the MC REFL PD transimpedance. See ELOG and related.

- ASC

Why do we want to see less PRM motion? I thought PRC motion was causing
LSC issue of the central part. We wanted to maximize the PRM effect, don't we?
(Or is this to supress ETM motion during full lock?)

 End PDH - good point, thanks.

ASC - Yes, this is so that we can use the POP QPD to feed back to the common ETMs after the CARM offset is already quite small.  We will not use POP DC QPD for PRC any more. 

Also, for future PRC ASC, I keep coming back to this in my head, but maybe it is less painful to install oplevs for PR2, PR3 than it would be to make an RF QPD.  Neither is going to be trivially easy.  But if we had sensors of the tip tilt motions, we could feed all of that back to the PRM to stabilize the PRC.

  10387   Thu Aug 14 18:02:11 2014 KojiUpdateGeneralUpdated game plan

Got the idea of ASC.

- Oplevs for PR2, PR3 => PR2 seems OK. PR3 almost impossible. well turned out not too crazy. We need outside electronics.

- RF QPD => not trivial and very technical but possible. All outside work.

- Better TT => might be a good solution.

  10388   Thu Aug 14 18:05:05 2014 JenneUpdateGeneralUpdated game plan

Quote:

- Oplevs for PR2, PR3 => Almost impossible.

 Because of the limited table space inside?  That's the main reason I can think of that this method is hard.  Am I missing something?

  10423   Fri Aug 22 13:51:00 2014 JenneUpdateGeneralUpdated game plan

40mToDoList.pdf

  6161   Tue Jan 3 17:46:38 2012 Updated Stewart platform design requirementsUpdateGeneralUpdated design requirements for a Stewart platform

I updated the design requirements for the Stewart platform.  I weighed the unloaded "dirty" SOS that was sitting on the workbench in the control room; its mass is 11 kg.  Steve suggested that the OSEMs (not installed on this model) would add another 0.5 kg.  From the specs in the final SOS design document, LIGO-T970135, I added 0.25 kg for the optic itself; I am therefore taking the total payload mass to be 11.75 kg.  (Now, the upper stage of the Stewart platform itself will likely add a nontrivial amount, but I am not worrying about this yet.)

I have e-mailed Janeen Romie to obtain the actual center of mass and principal moments of inertia of the platform.  I also cooked up a simple scheme to measure both quantities, should this information not be available.  It would involve rigidly mounting the dirty SOS to a rigid bar hung from a pivot.  By translating the mount point in two dimensions and measuring the period of the pendulum, I ought to be able to find the center of mass and moments of inertia by multilinear regression.  However, this elaborate scheme is not necessary to just compute some ballpark figures; it could wait until a later stage in the design.  For the time being, I just rescaled the moment of inertia proportional to the increase in mass, such that the torque-to-force ratio is unchanged.

As such, the design requirements are now 

  • linear travel: 40 microns peak to peak (based on SOS design requirements in LIGO-T950011)
  • angular travel: 3 mrad peak to peak (based on SOS design requirements in LIGO-T950011)
  • payload mass: 11.75 kg (measured unloaded mass, plus educated guesses about combined mass of OSEMs and optic)
  • payload moment of inertia: 0.0232 kg m^2 (wild guess)
  • bandwidth: 500 Hz (suggestion of Rana and Koji: ~kHz)

From these assumptions, the revised actuator requirements and dimensions are:

  • peak actuator force: 2.04 kN
  • minimum radius of top platform: 15 cm
  • minimum radius of bottom platform: 30 cm
  • minimum height: 26 cm

See the attached PDF document.

It appears that the actuator that I had originally nominated, PI's model P-225.80, would very nearly meet the actuator force requirement.  Steve also pointed out the following single-axis shakers that are already in use in the 40m:

  • Brüel & Kjær type 4809
  • Brüel & Kjær type 4810

I want to find out if either of these would meet the present need, but I'm waiting on a response from the manufacturer to get access to the data sheets.

 

  3715   Thu Oct 14 10:59:10 2010 josephbUpdateComputersUpdated cshrc.40m and Computer Restart Procedures

I started modifying the cshrc.40m file in /cvs/cds/caltech/ so that it starts pointing at the new directories.

# misc aliases
alias target 'cd /opt/rtcds/caltech/c1/target'
alias scripts 'cd /opt/rtcds/caltech/c1/scripts'
alias chans 'cd /opt/rtcds/caltech/c1/chans'
alias c 'cd /opt/rtcds/caltech/c1'
alias s 'cd /opt/rtcds/caltech/c1/scripts'
alias u 'cd /cvs/cds/caltech/users'

I also added "alias screens 'cd /opt/rtcds/caltech/c1/medm'" as a quick way to get to the medm directory.

Once we get multiple compiled versions (i.e. i386, x86_64, Solaris) of the new gds tools from Alex, we'll have to some more serious surgery on this file.

I removed the "New Computer Restart Procedures" and simply moved the changes into the "Computer Restart procedures", found here.  I've removed everything I don't think applies anymore (all the VME FE reboot procedures for example).

  4265   Wed Feb 9 15:26:22 2011 josephbUpdateCDSUpdated c1scx with lockin, c1gcv for green transmission pd

Updated the c1scx model to have two Lockin demodulators (C1:SUS-ETMX_LOCKIN1 and C1:SUS-ETMX_LOCKIN2).  There is a matrix C1:SUS-ETMX_INMUX which directs signals to the inputs of LOCKIN1 and LOCKIN2.  Currently only the GREEN_TRX signal is the only signal going in to this matrix, the other 3 are grounds.  The actual clocks themselves had to be at the top level (they don't work inside blocks) and thus named C1:SCX-ETMX_LOCKIN1_OSC and C1:SCX-ETMX_LOCKIN2_OSC.

 

There is a signal (IPC name is C1:GCV-SCX_GREEN_TRX) going from the c1gcv model to the c1scx model, which will contain the output from Jenne's green transmission PD which will eventually be placed. I've placed a filter bank on it in the c1gcv model as a monitor point, and it corresponds to C1:GCV-GREEN_TRX.

 

The suspension control screens were modified to have a screen for the Matrix feeding signals into the two lockin demodulators.  The green medm screen was also modified to have readbacks for the GREEN_TRX and GREEN_TRY channels.

 

So on the board, the top channel (labeled 1, corresponds to code ADC_0_0) is MCL.

Channel 2 (ADC_0_1) is assigned to frequency divided green signal.

Channel 3 (ADC_0_2) is assigned to the beat PD's DC output.

Channel 4 (ADC_0_3) is assigned to the green power transmission for the x-arm.

Channel 5 (ADC_0_4) is assigned to the green power transmission for the y-arm.

  4200   Tue Jan 25 15:20:38 2011 josephbUpdateCDSUpdated c1rfm model plus new naming convention for RFM/Dolphin

After sitting down for 5 minutes and thinking about it, I realized the names I had been using for internal RFM communication were pretty bad.  It was because looking at a model didn't let you know where the RFM connection was coming from or going to.  So to correct my previous mistakes, I'm instituting the following naming convention for reflected memory, PCIE reflected memory (dolphin) and shared memory names.  These don't actually get used anywhere but the models, and thus don't show up as channel names anywhere else.  They are replaced by raw hex memory locations in the actual code through the use of the IPC file (/opt/rtcds/caltech/c1/chans/ipc/C1.ipc).  However it will make understanding the models easier for anyone looking at them or modifying them.

 

The new naming convention for RFM and Dolphin channels is as follows.

SITE:Sending Model-Receiving Model_DESCRIPTION_HERE

The description should be unique to that data being transferred and reused if its the same data.  Thus if its transfered to another model, its easy to identify it as the same information.

The model should be the .mdl file name, not the subsystem its a part of.  So SCX is used instead of SUS.  This is to make it easier to track where data is going.

In the unlikely case of multiple models receiving, it should be of the form SITE:Sending Model-Receiving Model 1-Receiving Model 2_DESCRIPTION_HERE.  Seperate models by dashes and description by underscores.

Example:

C1:LSC-RFM_ETMX_LSC

This channel goes from the LSC model (on c1lsc) to the RFM model (on c1sus).  It transfers ETMX LSC position feedback.  The second LSC may seem redundant until we look at the next channel in the chain.

C1:RFM-SCX_ETMX_LSC

This channel goes from the RFM model to the SCX model (on c1iscex). It contains the same information as the first channel, i.e. ETMX LSC position feedback.

 

I have updated all the models that had RFM and SHMEM connections, as well as adding all the LSC communciation connections to c1rfm.  This includes c1sus, c1rfm, c1mcs, c1ioo, c1gcv, c1lsc, c1scx, c1scy.  I have not yet built all the models since I didn't finish the updates until this afternoon.  I will build and test the code tomorrow morning.

 

 

 

  16037   Thu Apr 15 17:24:08 2021 JonUpdateCDSUpdated c1auxey wiring plan

I've updated the c1auxey wiring plan for compatibility with the new suspension electronics. Specifically it is based on wiring schematics for the new HAM-A coil driver (D1100117), satellite amplifier (D1002818), and HV bias driver (D1900163).

Changes:

  • The PDMon, VMon, CoilEnable, and BiasAdj channels all move from DB37 to various DB9 breakout boards.
  • The DB9 cables (x2) connecting the CoilEnable channels to the coil drivers must be spliced with the dewhitening switching signals from the RTS.
  • As suggested, I added five new BI channels to monitor the state of the CoilEnable switches. For lack of a better name, they follow the naming convention C1:SUS-ETMY_xx_ENABLEMon.

@Yehonathan can proceed with wiring the chassis.

Quote:

I finished prewiring the new c1auxey Acromag chassis (see attached pictures). I connected all grounds to the DIN rail to save some wiring. The power switches and LEDs work as expected.

I configured the DAQ modules using the old windows machine. I configured the gateway to be 192.168.114.1. The host machine still needs to be setup.

Next, the feedthroughs need to be wired and the channels need to be bench tested.

  16052   Mon Apr 19 21:54:55 2021 YehonathanUpdateCDSUpdated c1auxey wiring plan

Except for the feed-throughs that require a DB9-M connector I finished wiring and labeling the Acromag, following Jon's updated wiring plan.

We can start testing the differential inputs until the missing connectors arrive.

 

  16092   Wed Apr 28 18:56:57 2021 Yehonathan, JonUpdateCDSUpdated c1auxey wiring plan

We took a Supermicro from the lab (along with a keyboard, a mouse, and a screen taken from a table on the Y arm) and placed it near the Acromag chassis.

We installed Debian 10 on the machine. I followed the steps on the slow machine wiki for setting up the host machine. Some steps had to be updated. Most importantly, in the new Debian, the network interfaces are given random names like enp3s0 and enp4s0 instead of eth0 and eth1. I updated the wiki accordingly.

To operate the chassis using one 15V source I disconnected the +24V cable from the Acromag units and jumpered the +15V wire into the power input instead. I started up the Acromags. They draw 0.7A. I connected an Ethernet cable to the front interface. I checked that all the Acromags are connected to the local network of the host machine by pinging them one by one.

 

  16098   Thu Apr 29 16:35:51 2021 YehonathanUpdateCDSUpdated c1auxey wiring plan

I installed the EPICs base, asyn and modbus modules according to Jon's instructions.

Since the modbus configurations files were already writtten for c1auxey1 (see elog 15292) the only thing I did was to change the IP addresses in ETMYaux.cmd to match the actual assigned IPs.

I followed the rest of the instructions as written.

The modbus service was activated succesfully.

The only thing left to do is to change ETMYaux.db to reflect to new channels that were added. I believe these are BI channels named C1:SUS-ETMY_xx_ENABLEMon.

 

  16103   Thu Apr 29 19:55:45 2021 YehonathanUpdateCDSUpdated c1auxey wiring plan

We received a stock of DB9 male feed-through connectors. That allowed me to complete the remaining wiring on the c1auxey Acromag chassis. The only thing left to be done is the splicing to the RTS.

 

  16107   Fri Apr 30 19:18:51 2021 Yehonathan, JonUpdateCDSUpdated c1auxey wiring plan

We finished the installation procedure on the c1auxey1 host machine. There were some adjustments that had to be made for Debian 10. The slow machine wiki page has been updated.

A test database file was made were all the channel names were changed from C1 to C2 in order to not interfere with the existing channels.

We starting testing the channels one by one to check the wiring and the EPICs software. We found some misswirings and fixed them.

Channel Name Type EPICs Test Acromag windows software test
C2:SUS-ETMY_ULPDMon AI Pass Pass
C2:SUS-ETMY_URPDMon AI Pass Pass
C2:SUS-ETMY_LLPDMon AI Pass Pass
C2:SUS-ETMY_SPDMon AI Pass Pass
C2:SUS-ETMY_LRPDMon AI Pass Pass
C2:SUS-ETMY_ULVMon AI Pass Pass
C2:SUS-ETMY_URVMon AI Pass Pass
C2:SUS-ETMY_LLVMon AI Pass Pass
C2:SUS-ETMY_SideVMon AI Pass Pass
C2:SUS-ETMY_LRVMon AI Pass Pass

Its getting late. I'll continue with the rest of the channels on Monday.

Notice that for all the AI channels the RTN was disconnected while testing.

 

 

 

 

 

 

  16114   Mon May 3 20:36:46 2021 Yehonathan, JonUpdateCDSUpdated c1auxey wiring plan

It seemed like the BIO channels were not working, both the inputs and the outputs. The inputs were working on the windows machine though. That is, when we shorted the BIO channel to the return, or put 0V on it, we could see the LED turn on on the I/O testing screen and when we ramped up the voltage above 3 the LED turned off. This is the expected behavior from a sinking digital input. However, the EPICs caget didn't show any change. All the channels were stuck on Disabled.

We checked the digital outputs by connecting the channels to a fluke. Initially, the fluke showed 13V. We tried to toggle the digital output channels with caput and that didn't work. We checked the outputs with the windows software. For that, we needed to stop the Modbus. To our surprise, the windows software was not able to flip the channels either. We realized that this BIO Acromag unit is probably defective. We replaced it with a different unit and put a warning sticker on the defective unit. Now, the digital outputs were working as expected. When we turned them on the voltage output dropped to 0V. We checked the channels with the EPICs software. We realized that these channels were locked with the closed loop definition. We turned on the channels tied to these output channels (watchdog and toggles) and it worked. The output channels can be flipped with the EPICs software. We checked all the digital output channels and fixed some wiring issues along the way.

The digital input channels were still not working. This is a software issue that we will have to deal with later.

(Yehonathan) Rana noticed that the BNC leads on the chassis front panel didn't have isolation on them so I redid them with shrinking tubes.

  2844   Mon Apr 26 11:29:37 2010 josephbUpdateComputersUpdated bitwise.pm in RCG SVN plus other fixes

To fix a problem one of the models was having, I checked the CVS version of the Bitwise.pm file into the SVN (located in /home/controls/cds/advLigoRTS/src/epics/util/lib), which adds left and right bit shifting funtionality.  The yec model now builds with the SVN checkout.

Also while trying to get things to work, I discovered the cdsRfmIO piece (used to read and write to the RFM card) now only accepts 8 bit offsets.  This means we're going to have to change virtually all of the RFM memory locations for the various channels, rather than using the values from its previous incarnation, since most were 4 bit numbers.  It also means it going to eat up roughly twice as much space, as far as I can tell.

Turns out the problem we were having getting to compile was nicely answered by Koji's elog post.  The shmem_daq value was not set to be equal to 1.  This caused it to look for myrimnet header files which did not exist, and caused compile time errors.  The model now compiles on megatron.

[Edit by KA: 4 bit and 8 bit would mean "bytes". I don't recall which e-log of mine Joe is referring.]

 

  3978   Tue Nov 23 16:55:14 2010 josephbUpdateCDSUpdated apps

Updated Apps:

I created a new setup script for the newest build of the gds tools (DTT, foton, etc), located in /opt/apps (which is a soft link from /cvs/cds/apps) called gds-env.csh.

This script is now sourced by cshrc.40m for linux 64 bit machines.  In addition, the control room machines have a soft link in the /opt directory to the /cvs/cds/apps directory.

So now when you type dtt or foton, it will bring up the Centos compiled code Alex copied over from Hanford last month.

  3983   Tue Nov 23 23:52:49 2010 ranaUpdateCDSUpdated apps

Wow. I typed DTT on rossa and it actually worked! No complaints about testpoints, etc. I was also able to use its new 'NDS2' function to get data off of the CIT cluster (L1:DARM_ERR from February). You have to use the kinit/kdestroy stuff to use NDS2 as usual (look up NDS2 in DASWG if you don't know what I mean).

  17122   Wed Aug 31 11:39:48 2022 YehonathanUpdateLSCUpdated XARM noise budget

{Radhika, Paco, Yehonathan}

For educational purposes we update the XARM noise budget and add the POX11 calibrated dark noise contribution (attachment).

  11395   Wed Jul 8 17:46:20 2015 JessicaSummaryGeneralUpdated Time Delay Plots

I re-measured the transfer function for Cable B, because the residuals in my previous post for cable B indicated a bad fit. 

I also realized I had made a mistake in calculating the time delay, and calculated more reasonable time delays today. 

Cable A had a delay of 202.43 +- 0.01 ns.

Cable B had a delay of 202.44 +- 0.01 ns. 

  13004   Mon May 22 15:01:41 2017 jigyasaUpdatetelescope designUpdated Telescope design with 1'' eye piece

I examined the use of a single lens system for the available range of focal lengths, for the required magnification and found that a focal length of at most 100 mm would be required to sufficiently cover the object distance range. This would greatly compromise with the f-number and hence lead to a lot more spherical aberrations.

Therefore, a two lens system would be more useful to implement. Using an eyepiece of 1” puts an additional constraint on the system such that the separation between the lenses must now at least equal or be greater than half the image distance from the first lens to ensure that no light from the light cone is lost. This is clarified in the schematic. The image from the first lens in absence of the second lens would form at point A, subtending an angle θ. In order to ensure that no part this light cone emerging from the first lens is lost, the second lens must be placed at a distance atleast v/2 from the first lens.

A combination of 125mm focal length 2” diameter objective with a 250 mm 1” eyepiece covers the required range of object distances (650mm to 1500 mm). Increasing the focal length of the eye piece increases the minimum object distance accessible to 700 mm. 

A glance at the accessible u, v points shows that all magnifications are not possible at a given object distance. To image the entire surface of the test mass, a distance of at least 1.25m is required from the objective, while a beam spot of 1'' diameter can be imaged easily at upto 1200 mm from the objective . This holds true even for the 150-250 mm biconvex 2" lens combination proposed earlier. 

If this sounds reasonable, we could proceed with ordering the lenses.

  3612   Mon Sep 27 17:35:13 2010 josephbUpdateCDSUpdated Suspension screens/Megatron now c1ioo/Further work on fb

The medm screens have been updated further, with the hidden matrices added in bright colors.  An example screen shot is attached.

Megatron has been renamed c1ioo and moved to martian network.  Similarly, c1sus and c1iscex are also on the martian network.  Medm screens can be run on any of the control machines and they will work.

Currently the suspension controller is running on c1sus.

The frame builder is currently running on the fb machine *however* it is not working well.  Test points and daq channels on the new front ends tended to crash it when Alex started the mx_stream to the fb via our new DAQ network (192.168.114.XXX, accessible through the front ends or fb - has a dedicated 1 gigabit network with up to 10 gigabit for the fb).  So for the moment, we're running without front end data. Alex will be back tomorrow to work on it. 

Alex claimed to have left the frame builder in a state where it should be recording slow data, however, I can't seem to access recent trends (i.e. since we started it today).  The frame builder throws up an error "Couldn't open raw minute trend file '/frames/trend/minute_raw/C1:Vac-P1_pressure', for example.  Realtime seems to work for slow channels however.  Remember to connect to fb, not fb40m. So it seems the fb is still in a mostly non-functional state.

Alex also started a job to convert all the old trends to the correct new data format, which should finish by tomorrow.

RA: Nice screen work. The old screens had a 'slow' slider effect when ramping the bias so that we couldn't whack the optic too hard. Is the new one instantaneous?

  3615   Tue Sep 28 10:07:29 2010 josephbUpdateCDSUpdated Suspension screens/Megatron now c1ioo/Further work on fb

Quote:

RA: Nice screen work. The old screens had a 'slow' slider effect when ramping the bias so that we couldn't whack the optic too hard. Is the new one instantaneous?

 Looking at the sliders, I apparently still need to connect them properly.  There's a mismatch between the medm screen channel name and the model name.  At the moment there is no "slow" slider effect implemented, so they are effectively instantaneous.  Talking with Alex, he suggests writing a little c-code block and adding it to the model.  I can use the c code used in the filter module ramps as a starting point.

  17134   Wed Sep 7 14:32:15 2022 AnchalUpdateSUSUpdated SD coil gain to keep all coil actuation signals digitally same magnitude

The coil driver actuation strength for face coils was increased by 13 times in LO1, LO2, SR2, AS1, AS4, PR2, and PR3.

After the change the damping strenghth of POS, PIT, and YAW were reduced, but not SIDE coil output filter module requires higher digital input to cause same force at the output. This wouldn't be an issue until we try to correct for cross coupling at output matrix where we would want to give similar order of magnitude coefficients for SIDE coil as well. So I did the following changes to make input to all coil output filters (Face and side) same for these particular suspensions:

  • C1:SUS-LO1_SUSSIDE_GAIN 40.0 -> 3.077
    C1:SUS-AS1_SUSSIDE_GAIN 85.0 -> 6.538
    C1:SUS-AS4_SUSSIDE_GAIN 41.0 -> 3.154
    C1:SUS-PR2_SUSSIDE_GAIN 150.0 -> 11.538
    C1:SUS-PR3_SUSSIDE_GAIN 100.0 -> 7.692
    C1:SUS-LO2_SUSSIDE_GAIN 50.0 -> 3.846
    C1:SUS-SR2_SUSSIDE_GAIN 140.0 -> 10.769
  • C1:SUS-LO1_SDCOIL_GAIN -1.0 -> -13.0
    C1:SUS-AS1_SDCOIL_GAIN 1.0 -> 13.0
    C1:SUS-AS4_SDCOIL_GAIN -1.0 -> -13.0
    C1:SUS-PR2_SDCOIL_GAIN -1.0 -> -13.0
    C1:SUS-PR3_SDCOIL_GAIN -1.0 -> -13.0
    C1:SUS-LO2_SDCOIL_GAIN -1.0 -> -13.0
    C1:SUS-SR2_SDCOIL_GAIN -1.0 -> -13.0

In short, now 10 cts of input to C1:SUS-AS1_ULCOIL would create same actuation on UL as 10 cts of input to C1:SUS-AS1_SDCOIL will on SD.


In reply to

Quote: http://nodus.ligo.caltech.edu:8080/40m/16802

[Koji, JC]

Coil Drivers LO2, SR2, AS4, and AS1 have been updated a reinstalled into the system. 

LO2 Coil Driver 1 (UL/LL/UR)now has R=100 // 1.2k ~ 92Ohm for CH1/2/3        Unit: S2100008

LO2 Coil Driver 2 (LR/SD)now has R=100 // 1.2k ~ 92Ohm for CH3                    Unit: S2100530

SR2 Coil Driver 1 (UL/LL/UR)now has R=100 // 1.2k ~ 92Ohm for CH1/2/3        Unit: S2100614

SR2 Coil Driver 2 (LR/SD)now has R=100 // 1.2k ~ 92Ohm for CH3                    Unit: S2100615

AS1 Coil Driver 1 (UL/LL/UR)now has R=100 // 1.2k ~ 92Ohm for CH1/2/3        Unit: S2100610

AS1 Coil Driver 2 (LR/SD)now has R=100 // 1.2k ~ 92Ohm for CH3                    Unit: S2100611

AS4 Coil Driver 1 (UL/LL/UR)now has R=100 // 1.2k ~ 92Ohm for CH1/2/3        Unit: S2100612

AS4 Coil Driver 2 (LR/SD)now has R=100 // 1.2k ~ 92Ohm for CH3                    Unit: S2100613

 

Quote: http://nodus.ligo.caltech.edu:8080/40m/16791

[JC Koji]

To give more alignment ranges for the SUS alignment, we started updating the output resistors of the BHD SUS coil drivers.
As Paco has already started working on LO1 alignment, we urgently updated the output Rs for LO1 coil drivers.
LO1 Coil Driver 1 now has R=100 // 1.2k ~ 92Ohm for CH1/2/3, and LO1 Coil Driver 2 has the same mod only for CH3. JC has taken the photos and will upload/update an elog/DCC.

We are still working on the update for the SR2, LO2, AS1, and AS4 coil drivers. They are spread over the workbench right now. Please leave them as they're for a while.
JC is going to continue to work on them tomorrow, and then we'll bring them back to the rack.

 

Quote: https://nodus.ligo.caltech.edu:8081/40m/16760

[Yuta Koji]

We took out the two coil driver units for PR3 and the incorrect arrangement of the output Rs were corrected. The boxes were returned to the rack.

In order to recover the alignment of the PR3 mirror, C1:SUS_PR3_SUSPOS_INMON / C1:SUS_PR3_SUSPIT_INMON / C1:SUS_PR3_SUSYAW_INMON were monitored. The previous values for them were {31150 / -31000 / -12800}. By moving the alignment sliders, the PIT and YAW values were adjusted to be {-31100 / -12700}. while this change made the POS value to be 52340.

The original gains for PR3 Pos/Pit/Yaw were {1, 0.52, 0.2}. They were supposed to be reduced to  {0.077, 0.04, 0.015}.
I ended up having the gains to be {0.15, 0.1, 0.05}. The side gain was also increased to 50.

----

Overall, the output R configuration for PR2/PR3 are summarized as follows. I'll update the DCC.

PR2 Coil Driver 1 (UL/LL/UR) / S2100616 / PCB S2100520 / R_OUT = (1.2K // 100) for CH1/2/3

PR2 Coil Driver 2 (LR/SD) / S2100617 / PCB S2100519 / R_OUT = (1.2K // 100) for CH3

PR3 Coil Driver 1 (UL/LL/UR) / S2100619 / PCB S2100516 / R_OUT = (1.2K // 100) for CH1/2/3

PR3 Coil Driver 2 (LR/SD) / S2100618 / PCB S2100518 / R_OUT = (1.2K // 100) for CH3

 

  2392   Thu Dec 10 18:27:48 2009 MottUpdateGeneralUpdated R & T Measurements

Attached are updated plots of the T&R Measurements for a variety of mirrors, and diagrams for the setup used to make the measurements.

T is plotted for the 1064 nm measurement, since these mirrors are highly reflective at 1064, and either R or R&T are plotted for the 532 nm measurement, depending on how larger the R signal is.

As with the previous set of plots, the error bars here are purely statistical, and there are certainly other sources of error not accounted for in these plots.  In general, the T measurement was quite stable, and the additional errors
are probably not enormous, perhaps a few percent.

The mirrors are:

Y1-1037-00.50CC

Y1-2037-45S

Y1-2037-45P

Y1S-1025-0

Y1S-1025-45

 

  2610   Wed Feb 17 12:45:19 2010 josephbUpdateComputersUpdated Megatron and its firewall

I updated the IP address on the Cisco Linksys wireless N router, which we're using to keep megatron separated from the rest of the network.  I then went in and updated megatrons resolv.conf and host files.  It is now possible to ssh into megatron again from the control machines.

  4808   Mon Jun 13 12:34:21 2011 Jamie, KojiUpdateLSCUpdated LSC model installed

After a couple of hickups, I was able to compile and install Koji's rework of the LSC model.

The main changes are that the model now use an RF_PD library part, and the channel names were tweaked to be more in line with what we expect.

I found a couple of small bugs in the model that were preventing it from compiling.  Those were fixed and it compiled with no further problems.

There was also some rearrangement of signal inputs to the PD_DOF matrix.  The matrix screen was updated to reflect the proper inputs.  However, this also meant that the burt restore scripts for the IFO configurations were setting the wrong elements in the matrix.  I fixed the settings for X and Y arm locking, and updated the burt snapshots using the burt/c1ifoconfigure/C1save{X,Y}arm scripts.  NOTE: burt settings will need to be updated for the MICH, PRM, DRM, and FULL IFO configurations as well.

During the build/install process, Joe and I also found a bug in the feCodeGen that was causing the filter screens to be created with the wrong names.  Joe sent out a patch that will hopefully be merged soon.  Building the model with Joe's patch fixed the screen names, so the screens are currently named correctly.

  14330   Tue Dec 4 10:38:12 2018 JonOmnistructureUpgradeUpdated Feedthrough List for Vacuum Acromag Chassis

Based on new input from Chub, attached is the revised list of signal cable feedthroughs needed on the vacuum system Acromag crate. I believe this list is now complete.

  913   Tue Sep 2 22:43:16 2008 YoichiConfigurationPSLUpdated FSS open loop TF
Since the LO level of the FSS servo was too low, I replaced the RF oscillator board with a combination of
a Stanford signal generator and an RF amplifier.
Right now, the POY RF amplifier is used for this purpose temporarily.
Now the LO level is about 16dBm. The RF power going into the EOM is attenuated by 20dB from the LO level.
I played with the cable length to get the phase right.
Then I was able to lock the FSS with the new RF signal source.

Attached is the open loop transfer function of the current FSS. Now the UGF is a bit above 200kHz, a factor of 2 improvement.
This gain was achieved with the common gain slider at 13.5dB and the fast gain = 30dB.
With the old RF oscillator board, UGF=100kHz was achieved with the common gain =30dB. Therefore, the increase of the LO gave
us a large signal gain.

Increasing the gain further, again ,makes the PC path crazy.
Rich suggested that this craziness was caused either by the slew rate limit of the PA85 or the output voltage limit of the bypass Op-amp(A829)
is hit.

TO DO:
* Look at the error signal spectrum to see if there is any signal causing the slew rate saturation at high frequencies.
* Find out what the RF signal level for the EOM should be. 20dB attenuation is an arbitrary choice.
* Find out the cross over frequency. Determine where the fast gain slider should be.
etc ...
  3962   Mon Nov 22 12:00:18 2010 josephbUpdateCDSUpdated Computer Restart Procedures for FB

I've updated the  Computer Restart Procedures  page in the wiki with the latest fb restart procedure.

To just restart just the daqd (frame builder) process, do:

1) telnet fb 8088

2) shutdown

The init process will take care of the rest and restart daqd automatically.

 

Background:
Plan:
  - Check the wiring after SOS Coil Driver Module and circuit around SDSEN
  - Check whitening and dewhitening filters. We connected a binary output cable, but didn't checked them yet.
  - Make a script for step 2
  - Activate new DAQ channels for ETMX (what is the current new fresh up-to-date latest fb restart procedure?)

 

  15738   Fri Dec 18 22:59:12 2020 JonConfigurationCDSUpdated CDS upgrade plan

Attached is the layout for the "intermediate" CDS upgrade option, as was discussed on Wednesday. Under this plan:

  • Existing FEs stay where they are (they are not moved to a single rack)

  • Dolphin IPC remains PCIe Gen 1

  • RFM network is entirely replaced with Dolphin IPC

Please send me any omissions or corrections to the layout.

  15742   Mon Dec 21 09:28:50 2020 JamieConfigurationCDSUpdated CDS upgrade plan
Quote:

Attached is the layout for the "intermediate" CDS upgrade option, as was discussed on Wednesday. Under this plan:

  • Existing FEs stay where they are (they are not moved to a single rack)

  • Dolphin IPC remains PCIe Gen 1

  • RFM network is entirely replaced with Dolphin IPC

Please send me any omissions or corrections to the layout.

I just want to point out that if you move all the FEs to the same rack they can all be connected to the Dolphin switch via copper, and you would only have to string a single fiber to every IO rack, rather than the multiple now (for network, dolphin, timing, etc.).

  15746   Wed Dec 23 23:06:45 2020 gautamConfigurationCDSUpdated CDS upgrade plan
  1. The diagram should clearly show the host machines and the expansion chassis and the interconnects between them.
  2. We no longer have any Gentoo bootserver or diskless FEs.
  3. The "c1lsc" host is in 1X4 not 1Y3.
  4. The connection between c1lsc and Dolphin switch is copper not fiber. I don't know how many Gbps it is. But if the switch is 10 Gbps, are they really selling interface cables that have lower speed? The datasheet says 10 Gbps.
  5. The control room workstations - Debian10 (rossa) is the way forward I believe. it is true pianosa remains SL7 (and we should continue to keep it so until all other machines have been upgraded and tested on Debian 10).
  6. There is no "IOO/OAF". The host is called "c1ioo".
  7. The interconnect between Dolphin switch and c1ioo host is via fiber not copper.
  8. It'd be good to have an accurate diagram of the current situation as well (with the RFM network).
  9. I'm not sure if the 1Y1 rack can accommodate 2 FEs and 2 expansion chassis. Maybe if we clear everything else there out...
  10. There are 2 "2GB/s" Copper traces. I think the legend should make clear what's going on - i.e. which cables are ethernet (Cat 6? Cat 5? What's the speed limitation? The cable? Or the switch?), which are PCIe cables etc etc. 

I don't have omnigraffle - what about uploading the source doc in a format that the excellent (and free) draw.io can handle? I think we can do a much better job of making this diagram reflect reality. There should also be a corresponding diagram for the Acromag system (but that doesn't have to be tied to this task). Megatron (scripts machine) and nodus should be added to that diagram as well.

Please send me any omissions or corrections to the layout.

  15771   Tue Jan 19 14:05:25 2021 JonConfigurationCDSUpdated CDS upgrade plan

I've produced updated diagrams of the CDS layout, taking the comments in 15476 into account. I've also converted the 40m's diagrams from Omnigraffle ($150/license) to the free, cloud-based platform draw.io. I had never heard of draw.io, but I found that it has most all the same functionality. It also integrates nicely with Google Drive.

Attachment 1: The planned CDS upgrade (2 new FEs, fully replace RFM network with Gen 1 Dolphin IPC)
Attachment 2: The current 40m CDS topology

The most up-to-date diagrams are hosted at the following links:

Please send me any further corrections or omissions. Anyone logged in with LIGO.ORG credentials can also directly edit the diagrams.

  15772   Tue Jan 19 15:43:24 2021 gautamConfigurationCDSUpdated CDS upgrade plan

Not sure if 1Y1 can accommodate both c1sus2 and c1bhd as well as the various electronics chassis that will have to be installed. There may need to be some distribution between 1Y1 and 1Y3. Does Koji's new wiring also specify which racks hold which chassis?

Some minor improvements to the diagram:

  1. The GPS receiver in 1X7 should be added. All the timing in the lab is synced to the 1pps from this.
  2. We should add hyperlinks to the various parts datasheets (e.g. Dolphin switch, RFM switch, etc etc) so that the diagram will be truly informative and self-contained.
  3. Megatron and nodus, but especially chiara (NFS server), should be added to the diagram. 
  11115   Sat Mar 7 19:23:27 2015 JenneUpdateLSCUpdated ALSwatch script

Last report on model change / screen work from yesterday.

The ALSwatch script has always been just looking at the EPICS output of the CARM and DARM filter banks, and if it saw a single saturation, it would run the down script.  This was non-ideal because (a) the EPICS channels aren't the real signals, and (b) sometimes we'll hit the rails briefly and that's okay - we want to shut things down only when we're constantly saturating.

It turns out that there was a pre-existing saturation monitor part in CDS_PARTS, which I have used.  There is one each looking at the output of the CARM and DARM filter banks.  The threshold for what saturation means is set as an epics input.  The part outputs a running count (number of saturations since the last time it was not saturated, resets each time it goes non-saturating) and a total number since the last reset (also an epics input). 

(To be continued... still writing)

  1205   Mon Dec 29 18:01:07 2008 AidanUpdateAuxiliary lockingUpdated 40m Upgrade Document T080074-00-R

Added a paragraph to the 40m Upgrade document describing the fiber stabilization and frequency doubling proposed for auxiliary locking.

Also added a complete diagram of the fiber stabilization and a draft sketch of the frequency doubling.

Uploaded to https://nodus.ligo.caltech.edu:30889/svn/trunk/docs/upgrade08/ via svn.
  5265   Thu Aug 18 22:24:08 2011 jamieOmnistructureVIDEOUpdated 'videoswitch' script

I have updated the 'videoswitch' program that controls the video MUX.  It now includes the ability to query the video mux for the channel mapping:

controls@pianosa:~ 0$ /opt/rtcds/caltech/c1/scripts/general/videoswitch -h
Usage:
videoswitch [options] [OUT]      List current output/input mapping [for OUT]
videoswitch [options] OUT IN     Set output OUT to be input IN

Options:
  -h, --help            show this help message and exit
  -i, --inputs          List input channels and exit
  -o, --outputs         List output channels and exit
  -l, --list            List all input and output channels and exit
  -H HOST, --host=HOST  IP address/Host name
controls@pianosa:~ 0$

  6956   Wed Jul 11 09:48:24 2012 LizSummaryComputer Scripts / ProgramsUpdate/daily summary testing

I have been working on configuration of the Daily Summary webpages and have been attempting to create a "PSL health" page.  This page will display the PMC power, the temperature on the PSL table and the PSL table microphone levels.  Thus far, I have managed to make the extra PSL tab and configure the graph of the interior temperature, using channel C1:PSL-FSS_RMTEMP.

I have been attempting to make a spectrogram for one of the PMC channels, but there is an issue with the spectrogram setup, as Duncan Macleod noted in ELOG 6686:

"At the moment a package version issue means the spectrogram doesn't work, but the spectrum should. At the time of writing, to use the spectrum simple add 'plot-dataplot2'."

 

 

Because of this issue, I have also been trying to make the spectrogram plots work.  Thus far, I have fixed the issue with one of the spectrogram plots, but there are several problems with the other four that I need to address.  I have also been looking at the microphone channels and trying to make the plot for them work.  I checked which microphone was on the PSL table and plotted it in matplotlib to make sure it was working.  However, when I tried to incorporate it into the daily summary pages, the script stops at that point!  It might simply be taking an excessively long time, but I have to figure out why this is the case.

                                                 (I am using channel C1:PEM-MIC_6_IN1_DQ, if this is blatantly wrong, please let me know!!)

 

The main point of this ELOG is that I have working test-daily summary pages online!  They can be found here:

https://nodus.ligo.caltech.edu:30889/40m-summary-test/archive_daily/20120710/

Also, if anyone has more requests for what they would like to see on the finalized summary pages site, please respond to this post or email me at: endavison@umail.ucsb.edu

  3085   Fri Jun 18 13:42:52 2010 KojiHowToGeneralUpdate your work

All SURFs (and all others as always) are supposed to post the update of your status on the elog.

In fact, I already heard that Sharmila had been working on the serial connection to TC-200 and made some results. All of us like to hear the story.

  5408   Wed Sep 14 20:04:05 2011 jamieUpdateCDSUpdate to frame builder wiper.pl script for GPS 1000000000

I have updated the wiper.pl script (/opt/rtcds/caltech/c1/target/fb/wiper.pl) that runs on the framebuilder (in crontab) to delete old frames in case of file system overloading.  The point of this script is to keep the file system from overloading by deleting the oldest frames.  As it was, it was not properly sorting numbers which would have caused it to delete post-GPS 1000000000 frames first.  This issue was identified at LHO, and below is the patch that I applied to the script.


--- wiper.pl.orig  2011-04-11 13:54:40.000000000 -0700
+++ wiper.pl       2011-09-14 19:48:36.000000000 -0700
@@ -1,5 +1,7 @@
 #!/usr/bin/perl
 
+use File::Basename;
+
 print "\n" .  `date` . "\n";
 # Dry run, do not delete anything
 $dry_run = 1;
@@ -126,14 +128,23 @@
 
 if ($du{$minute_trend_frames_dir} > $minute_frames_keep) { $do_min = 1; };
 
+# sort files by GPS time split into prefixL-T-GPS-sec.gwf
+# numerically sort on 3rd field
+sub byGPSTime {
+    my $c = basename $a;
+    $c =~ s/\D+(\d+)\D+(\d+)\D+/$1/g;
+    my $d = basename $b;
+    $d =~ s/\D+(\d+)\D+(\d+)\D+/$1/g;
+    $c <=> $d;
+}
+
 # Delete frame files in $dir to free $ktofree Kbytes of space
 # This one reads file names in $dir/*/*.gwf sorts them by file names
 # and progressively deletes them up to $ktofree limit
 sub delete_frames {
        ($dir, $ktofree) = @_;
        # Read file names; Could this be inefficient?
-       @a= <$dir/*/*.gwf>;
-       sort @a;
+       @a = sort byGPSTime <$dir/*/*.gwf>;
        $dacc = 0; # How many kilobytes we deleted
        $fnum = @a;
        $dnum = 0;
@@ -145,6 +156,7 @@
          if ($dacc >= $ktofree)  { last; }
          $dnum ++;
          # Delete $file here
+         print "- " . $file . "\n";
          if (!$dry_run) {     
            unlink($file);
          }