40m QIL Cryo_Lab CTN SUS_Lab TCS_Lab OMC_Lab CRIME_Lab FEA ENG_Labs OptContFac Mariner WBEEShop
  40m Log, Page 81 of 339  Not logged in ELOG logo
ID Date Author Type Category Subjectup
  10240   Sat Jul 19 01:59:34 2014 HarryUpdateGeneralFiber Mode Measurement

Purpose 

We wanted to measure the mode coming out of the fibers, so we can later couple it to experimental setups for measuring different noise sources within the fiber. i.e. Polarization Extinction Ratio, Frequency Noise, Temperature Effects.

Methods

I used the beamscan mounted on a micrometer stage in order to measure the spot sizes of the fiber coupled light at different points along the optical axis, in much the same way as in the razorblade setup I used earlier in the summer.

fiberModeMeasurement.png

Analysis

I entered my data (z coordinates, spot size in x, spot size in y) into a la mode to obtain the beam  profile (waist size, location)

 fiberModeMeasurement1.png 

Code is attached in .zip file.

Moving Forward

After I took these measurements, Manasa pointed out that I need points over a longer distance. (These were taken over the range of the micrometer stage, which is 0.5 inches.)

I will be coming in to the 40m early on Monday to make these measurements, since precious beamscan time is so elusive.

Eventually, we will use this measurement to design optical setups to characterize Polarization Extinction Ratio, Frequency Noise, and temperature effects of the fibers, for further use in FOL.

  10244   Mon Jul 21 10:30:38 2014 HarryUpdateGeneralFiber Mode Measurement

 Purpose

The idea was to measure the profile of the light coming out of the fiber, so we could have knowledge of it for further design of measurement apparatuses, for characterization of the fibers' properties.

Methods

The method was the same as the last time I tried to measure the fiber mode.

This time I moved the beam profiler in a wider range along the z-axis.

Additionally, I adjusted the coupling until it gave ~1mW through the fiber, so the signal was high enough to be reliably detectable.

Measurements were taken in both X and Y transections of the beam.

The range of movement was limited by the aperture of the beam profiler, which cuts off at 9mm. My measurements stop at 8.3mm, as the next possible measurement was beyond the beam profiler's range.

fiberModeMeasurement.png

Analysis

I entered my data into A La Mode, which gave me a waist of 5um, at a location of z = -0.0071 m, that is to say, 7.1mm inside the fiber.

Note that in the plot, data points and fits overlap, and so are sometimes hard to distinguish from each other.

Code is attached.

fiberModeFit2.png

Moving Forward

Using this data, I will begin designing setups to measure fiber characteristics, the first of which being Polarization Extinction Ratio.

Eventually, the data collected from these measurements will be put to use in the frequency offset locking setup.

  10249   Mon Jul 21 18:08:19 2014 HarryUpdateGeneralFiber Mode Measurement

Quote:

 Purpose

The idea was to measure the profile of the light coming out of the fiber, so we could have knowledge of it for further design of measurement apparatuses, for characterization of the fibers' properties.

Methods

The method was the same as the last time I tried to measure the fiber mode.

This time I moved the beam profiler in a wider range along the z-axis.

Additionally, I adjusted the coupling until it gave ~1mW through the fiber, so the signal was high enough to be reliably detectable.

Measurements were taken in both X and Y transections of the beam.

The range of movement was limited by the aperture of the beam profiler, which cuts off at 9mm. My measurements stop at 8.3mm, as the next possible measurement was beyond the beam profiler's range.

fiberModeMeasurement.png

Analysis

I entered my data into A La Mode, which gave me a waist of 5um, at a location of z = -0.0071 m, that is to say, 7.1mm inside the fiber.

Note that in the plot, data points and fits overlap, and so are sometimes hard to distinguish from each other.

Code is attached.

fiberModeFit2.png

Moving Forward

Using this data, I will begin designing setups to measure fiber characteristics, the first of which being Polarization Extinction Ratio.

Eventually, the data collected from these measurements will be put to use in the frequency offset locking setup.

 Edit

 

 

The previous data were flawed, in that they were taken in groups of three, as I had to move the micrometer stage which held the beamscan between holes in the optical table.

In order to correct for this, I clamped a straightedge (ruler) to the table, so I could more consistently align the profiler with the beam axis.

These data gave a waist w_o = 4um, located 6mm inside the fiber. While these figures are very close to what I would expect (3.3um at the end of the fiber) the fitting still isn't as good as I would like.

The fit given by ALM is below.

fiberModeMeasurement3.png

Moving Forward

I would like to get a stage//rail so I can align the axes of the beam and profiler more consistently.

I would also like to use an aperture the more precisely align the profiler aperture with the beam axis.

Once these measurements have been made, I can begin assembling the setup to measure the Polarization Extinction Ratio of the fiber.

  10255   Tue Jul 22 16:26:04 2014 HarryUpdateGeneralFiber Mode Measurement

I repeated this process once more, this time using the computer controlled stage that the beam profiler is designed to be mounted to.

These data//fitting appears to be within error bars. The range of my measurements was limited when the beam width was near the effective aperture of the profiler.

This latest trial yielded a waist of 4um, located 2.9 mm inside the fiber for the X profile, and 3.0mm inside the fiber for the Y profile.

fiberModeProfile3.png

Code is attached in fiberModeMeasurement4.zip. Note that the z=0 point is defined as the end of the fiber.

  10282   Mon Jul 28 17:25:32 2014 HarryUpdateGeneralFiber Mode With Collimators

 Purpose

We want a measurement of the fiber modes at either end, with the collimators, because these will be the modes that we'll be trying to match in order to couple light into the fibers, for FOL and/or future projects.

Measurement

In order to measure these modes, I used the beam profiler (Thorlabs BP 209-VIS) to take measurements of the beam diameter (cut off at 13.5% of the amplitude) along the optical axis, for each of the fiber ends.

The ends are arbitrarily labelled End 1 and End 2.

For each measurement, the fibers were coupled to roughly 30%, or 25mW at the output.

Regarding the issue of free rotation in the collimator stages: while End 1 was relatively stable, End 2 tended to move away from its optimal coupling position. In order to correct for this, I chose a position where coupling was good, and repositioned the stage to that coordinate (124 degrees) before taking each measurement.

The data were then entered into A La Mode, which gave waist measurements as follows:

End 1--- X Waist: 197um at Z = 4.8mm       Y Waist: 190um at Z = 13.6mm

End 2--- X Waist: 192um at Z = 7.4mm       Y Waist: 190um at Z = 6.0mm

end1Profiles.pngend2Profiles.png

A La Mode code is attached in .zip file

Moving Forward

These are the types of profiles that we will hopefully be matching the PSL and AUX lasers to, for use in frequency offset locking.

More characterization of the fibers is to follow, including Polarization Extinction Ratio.

We also hope to be testing the overall setup soon.

 

 

  10985   Fri Feb 6 18:06:18 2015 manasaUpdateGeneralFiber Optic module for FOL

I pulled out the Fiber Optic Module for FOL from the rack inside the PSL table enclosure and modified it. The beat PDs were moved into the box to avoid breaking the fiber pigtail input to the PD.

The box has 3 input FC/APC connectors (PSL and AUX lasers) and 2 output FC/APC connectors (10% of the beatnote for the AUX lasers).

Attachment shows what is inside the box. The box will again go back on the rack inside the PSL enclosure.

  10403   Fri Aug 15 17:24:44 2014 HarryUpdateGeneralFiber Temp.

 Earlier today Q and I somewhat resurrected my old PER measurement setup so I could run the temperature characterization experiment.

Unfortunately, when I tried to use the fiber illuminator, no light came from the other end, causing me to fail my primary goal for the summer of "don't break anything." The fiber has been re-spooled and labeled appropriately. Also sorry.

In addition to this, Q and I scavenged parts from the telescopes on the PSL and Y End tables, which were either not functional, or needed to have their mode matching adjusted, since we're using the non-PM fibers for FOL, which have a different numerical aperture, and thus slightly different output modes.

Specifically, this is involved removing the rotational mounts, and appropriate beam dumping.

My "calorimeter" still remains intact, in case anyone wants to make this measurement in the future, as this is my last day in the lab.

It's also effective at keeping drinks cold, if you'd rather use it for that.

  10389   Thu Aug 14 18:10:46 2014 HarryUpdateGeneralFiber Temperature Effects Setup

Purpose

We want to characterize the sort of response the fibers have to temperature gradients along them (potentially altering indices of refraction, etc.)

Experimental Setup

I have constructed a sort of two chambered "calorimeter" (by which I mean some coolers and other assorted pieces of recycling.)

The idea is that half of the length of PM fiber resides in one chamber, and the other in the other.

One chamber will remain at an uncontrolled, stable temperature (as measured by thermocouple probe) while the other's temperature is varied using a heat gun.

Using this setup, one can measure losses in power, and effects on polarization within the fiber.

Caveat

This is currently living on the electronics bench until tomorrow morning, and is a little fragile, just in case it needs to be moved.

  14643   Wed May 29 18:13:25 2019 gautamUpdateALSFiber beam-splitters are now PM

To maintain PM fibers all the way through to the photodiode, I had ordered some PM versions of the 50/50 fiber beamsplitters from AFW technologies. They arrived some days ago, and today I installed them in the BeatMouth. Before installation, I checked that the ends of the fibers were clean with the fiber microscope. I also did a little cleanup of the NW corner of the PSL table, where the 1um MZ setup was completely disassembled. We now have 4 non-PM fiber beamsplitters which may be useful for non polarizaiton sensitive applications - they are stored in the glass-door cabinet slightly east of the IY chamber along the Y arm, together with all the other fiber-related hardware.

Anjali had changed the coupling of the beam to the slow axis for her experiment but I ordered beamsplitters which have the slow axis blocked (because that was the original config). I need to revert to this config, and then make a measurement of the ALS noise - if things look good, I'll also patch up the Y arm ALS. We made several changes to the proposed timeline for the summer but I'd like to see this ALS thing through to the end while I still have some momentum before embarking on the BHD project. More to follow later in the eve.

Quote:

Get a fiber BS that is capable of maintaining the beam polarization all the way through to the beat photodiode. I've asked AFW technologies (the company that made our existing fiber BS parts) if they supply such a device, and Andrew is looking into a similar component from Thorlabs.

  14503   Sun Mar 31 15:05:53 2019 gautamUpdateALSFiber beam-splitters not PM

I looked into this a little more today.

  1. Looking at the beat signal between the PSL and EX beams from the NF1611 on a scope (50-ohm input), the signal Vpp was ~200 mV.
  2. In the time that I was poking about, the level dropped to ~150mVpp. seemed suspicious.
  3. Thinking that this has to be related to the polarization mismatch between the interfering beams, I moved the input fibers (blue in Attachment #1) around, and saw the signal amplitude went up to 300mVpp, supporting my initial hypothesis.
  4. The question remains as to where the bulk of the polarization drift is happening. I had spent some effort making sure the input coupled beam to the fiber was well-aligned to one of the special axes of the fiber, and I don't think this will have changed since (i.e. the rotational orientation of the fiber axes relative to the input beam was fixed, since we are using the K6XS mounts with a locking screw for the input couplers). So I flexed the patch cables of the fiber beam splitters inside the BeatMouth, and saw the signal go as high as 700mVpp (the expected level given the values reported by the DC monitor).

This is a problem - such large shifts in the signal level means we have to leave sufficient headroom in the choice of RF amplifier gain to prevent saturation, whereas we want to boost the signal as much as possible. Moreover, this kind of operation of tweaking the fiber seating to increase the RF signal level is not repeatable/reliable. Options as I see it:

  1. Get a fiber BS that is capable of maintaining the beam polarization all the way through to the beat photodiode. I've asked AFW technologies (the company that made our existing fiber BS parts) if they supply such a device, and Andrew is looking into a similar component from Thorlabs.
    • These parts could be costly.
  2. Mix the beams in free space. We have the beam coming from EX to the PSL table, so once we mix the two beams, we can use either a fiber or free-space PD to read out the beatnote. 
    • This approach means we lose some of the advantages of the fiber based setup (e.g. frequent alignment of the free-space MM of the two interfering beams may be required).
    • Potentially increases sensitivity to jitter noise at the free-space/fiber coupling points
Quote:
    • An initial RF beat power level measurement yielded -5dBm, which is inconsistent with the DC monitor voltages, but I'm not sure what frequency the beat was at, will make a more careful measurement with a scope or the network analyzer.
  11013   Thu Feb 12 12:16:04 2015 manasaUpdateGeneralFiber shielding

[Steve, Manasa]

The fibers around the PSL table were shielded to avoid any tampering.

  20   Fri Oct 26 21:48:40 2007 waldmanConfigurationOMCFiber to 056
I set up a 700 mW NPRO in Rana's lab and launched it onto a 50m fiber. I got a few mW onto the fiber, enough to see with a card before disabling the laser. The fiber now runs along the hallway and terminates in rm 056. Its taped down everywhere someone might trip on it, but don't go out of your way to trip on it or pull on it because you are curious. Tomorrow I will co-run a BNC cable and attenuate the NPRO output so it can only send a few mW and so be laser safe. Then we can try to develop a procedure to align the beam to a suspended OMC and lock our suspended cavity goodness.

Notes to self: items needed from the 40m
  • ND10 and ND20 neutral density filter
  • EOM and mount set for 4 inch beam height
  • Post for fiber launch to get to 4 inch
  • Mode matching lens at 4in
  • 3x steering mirror at 4in
  • RF photodiode at 4in
  • Post for camera to 4in
  • Light sheild for camera
  • Long BNC cable
Some of these exist at 056 already
  3594   Wed Sep 22 16:35:45 2010 josephbUpdateCDSFibers pulled, new FB install tomorrow

[Aidan, Tara, Joe]

We pulled out what used to be the LSC/ASC fiber from the 1Y3 arm rack, and then redirected it to the 1X1 rack.  This will be used as the c1ioo 1PPS timing signal.  So c1ioo is using the old c1iovme fiber for RFM communications back to the bypass switch, and the old LSC fiber for 1PPS.

The c1sus machine will be using the former sosvme fiber for communications to the RFM bypass switch.  It already had a 1 PPS timing fiber.

The c1iscex machine had a new timing fiber already put in, and will be using the c1iscey vme crate's RFM for communication.

We still need to pull up the extra blue fiber which was used to connect c1iscex directly to c1sus, and reuse it as the 1PPS signal to the new front end on the Y arm. 

Alex has said he'll come in tomorrow morning to install the new FB code.

 

  13753   Fri Apr 13 17:56:26 2018 gautamUpdateALSFibers switched out

I swapped the EX fiber for the PSL fiber in the polarization monitoring setup. There is a lot more power in this fiber, and one of the PDs was saturated. I should really have put a PBS to cut the power, but I opted for putting an absorptive ND1.0 filter on the PD instead for this test. I want to monitor the stability in this beam and compare it to the EX beam's polarization wandering.

  13754   Sat Apr 14 14:42:09 2018 gautamUpdateALSFibers switched out

It looks like the drift in polarization content in the PSL pickoff is actually much higher than that in the EX pickoff. Note that to prevent the P-pol diode from saturating, I put an ND filter in front of the PD, so the Y axis actually has to be multiplied by 10 to compare power in S and P polarizations. If this drift is because of the input (linear) polarization being poorly matched to one of the fiber's special axes, then we can improve the situation relatively easily. But if the polarization drift is happening as a result of time-varying stress (due to temp. fluctuations, acoustics etc) on the (PM) fiber from the PSL fiber coupler to the BeatMouth, then I think this is a much more challenging problem to solve.

I'll attempt to quantify the contribution (in Hz/rtHz) of beat amplitude RIN to phase tracker output noise, which will tell us how much of a problem this really is and in which frequency bands. In particular, I'm interested in seeing if the excess noise around 100 Hz is because of beat amplitude fluctuations. But on the evidence thus far, I've seen the beat amplitude drift by ~15 dB (over long timescales) on the control room network analyzer, and this drift seems to be dominated by PSL light amplitude fluctuations.

  13757   Tue Apr 17 14:08:29 2018 gautamUpdateALSFibers switched out

A follow-up on the discussion from today's lunch meeting - Rana pointed out that rotation of the fiber in the mount by ~5degrees cannot account for such large power fluctuations. Here is a 3 day trend from my polarization monitoring setup. Assuming the output fiber coupler rotates in its mount by 5 degrees, and assuming the input light is aligned to one of the fiber's special axes, then we expect <1% fluctuation in the power. But the attached trend shows much more drastic variations, more like 25% in the p-polarization (which is what I assume we use for the beat, since the majority of light is in this polarization, both for PSL and EX). I want to say that the periodicity in the power fluctuations is ~12hours, and so this fluctuation is somehow being modulated by the lab temperature, but unfortunately, we don't have the PSL enclosure temperature logged in order to compare coherence.

Steve: your  plots look like temperature driven


The "beat length" of the fiber is quoted as <=2.7mm. This means that a linearly polarized beam that is not oriented along one of the special axes of the fiber will be rotated through 180 degrees over 2.7mm of propagation through the fiber. I can't find a number for the coefficient of thermal expansion of the fiber, but if temperature driven fluctuations are changing the length of the fiber by 300um, it would account for ~12% power fluctuation between the two polarizations in the monitoring setup, which is in the ballpark we are seeing...

  13773   Fri Apr 20 00:26:34 2018 gautamUpdateALSFibers switched out

Summary:

I think the dominant cause for the fact that we were seeing huge swing in the power coupled into the fiber was that the beam being sent in was in fact not linearly polarized, but elliptically polarized. I've rectified this with the help of a PBS. Fiber has been plugged into my polarization monitoring setup. Let's monitor for some long stretch and see if the situation has improved.

Details:

  • The new fiber mount I ordered, K6XS, arrived today. I like it - it has little keys with which all DoFs can be locked. Moreover, it is compatible with the fixed collimators which IMO is the easiest way to achieve good mode-matching into the fiber. It is basically a plug-and-play replacement for the mounts we were using. Anyways, we can evaluate the performance over the coming days.
  • I installed it on the PSL table (started work ~10pm, HEPA turned up to maximum, PSL shutter closed).
  • But even with the new rotational DoF locking feature, I saw that slight disturbances in the fiber caused wild fluctuations in my polarization monitoring setup PD outputs. This was a useful tool through the night of checking the polarization content in the two special axes - Aidan had suggested using a heat gun but shaking the fiber a bit works well too I think.
  • The PM980 fiber has an alignment key that is aligned with the slow axis of the fiber - so it is a useful alignment reference. But even by perturbing the roational alignment about the vertical by +/-15 degrees, I saw no improvement in this behavior. So I began to question my assumption that the input beam itself had clean polarization content.
  • Since my pickoff beam has gone through a QWP and two PBSs, I had assumed that the beam was linearly polarized.
  • But by putting a PBS just upstream of the input fiber coupler, I could see a beam at the S-port with an IR card (while I expected the beam to be P-polarized).
  • OK - so I decided to clean up the input polarization by leaving this PBS installed. With this modification to the setup, I found that me shaking the fiber around on the PSL table didn't affect the output polarization content nearly as dramatically as before!!yes
  • The state I am leaving it in tonight is such that there is ~100x the power in the P-polarization output monitor as the S-polarization (PER ~ 20dB). I didn't try and optimize this too much more for now, I want to observe some long term trend to see if the wild power fluctuations have been mitigated.
  • The output coupler is mounted on the inferior K6X mount, and so there is the possibility that some drift will be attributable to rotation of the output coupler in it's mount. Thermally driven length changes / time varying stresses in the fiber may also lead to some residual power fluctuations. But I don't expect this to be anywhere near the ~25% I reported in the previous elog.
  • The rejected beam from the PBS was measured to be ~300 uW. I haven't dumped this properly, to be done tomorrow.
  • HEPA turned back down to 30%, PSL enclosure closed up, PSL shutter re-opened ~0030am.
  • Note that the EX and EY fiber coupled beams are also likely subject to the same problem. We have to double check. I think it's better to have a PBS in front of the input fiber coupler as this also gives us control over the amount of light coupled into the fiber.

Power budget:

Power in Measured power (Ophir, filter OFF)
@Input coupler, before PBS 4.4 mW
P-pol content @ input coupler 4.06 mW
S-pol (rejected) from PBS 275 uW
@Output coupler 2.6 mW (MM ~65%)

 

  4882   Sat Jun 25 00:00:28 2011 SonaliUpdateGreen LockingFibre Coupling.

 What I did today.

1. I tried to align the IR input beam by aligning the two mirrors, to couple input light into the fibre.

2.I was unsuccessful for a long time even though I tried a lot of tricks.

3. I also tried to use the optical fault locator to superpose the IR beam spot onto the beam spot of the other laser to facilitate effective coupling.

4.But the crucial point was to superpose the input beam path in the perfect direction of the output beam path and not just the beam spot.(the input cone and the output cone are perfectly aligned).

5.After one whole day of trial and thought, I managed to couple light into the fibre, and saw the output beam spot on the screen-camera-monitor set-up which we had arranged. Eurekka !!;)

6.I then used a power meter to measure the input beam power and the output beam power.

7.It was a disappointing 2% . I had read in project reports of many students of a 20% success.

8.After a lot of subtle tweaking of the mirrors using the knobs, I managed to increase the percentage of output beam to 12%.

9. This is a workable level.

10.A day of lot of new learning! Pictures of the setup are attached.:)

 

  3859   Thu Nov 4 03:13:46 2010 SureshUpdateLockingFibre coupling 1064nm light at the south-end table

[Kiwamu, Suresh]

We decided to use the 1064nm beam reflected from the Y1-1037-45-P mirror after the collimation lens following the doubling crystal for coupling into the optical fiber (ref 3843 and 3847 ).

We replaced a beam dump which was blocking this beam with a Y1-1037-45-P mirror and directed the beam into the fiber coupler with another Y1-1037-45-P.  The power in this beam was about 1W.  This has been stepped down to 10mW by introducing a reflective ND filter of OD=2.  The reflected power has been dumped into a blade-stack beam dump.

Steve has ordered the The Visual Fault Locator from Fluke.  It is expected to arrive within a day or two.

 

 

  3865   Thu Nov 4 19:00:57 2010 SureshUpdateLockingFibre coupling 1064nm light at the south-end table

The Fluke Visual Fault locator (Visifault) arrived and I used it to couple 1064nm light into the single mode fibre at the south-end-table.

Procedure used:

When the output end of the fiber is plugged into the Visifault the light emerges from at the south end (input side for 1064nm).  This light is collimated with the fiber coupler at that end and serves as a reference for the optical axis along which the 1064 light must be directed.  Once the two beams (red and 1064) are overlapped with the beam steering mirrors, the Visifault was disconnected from the fiber and the  fibre output ( 1064 at the PSL table) is maximized by walking the beam at the input end and adjusting the collimation at the input.

The output of the fiber has been collimated with a fiber coupler.

7.5mW are incident on the input end and 1.3mW have been measured at the output.    This output power is adequate for the observing the beats with PSL NPRO.

 

 

 

  8183   Wed Feb 27 14:39:59 2013 AnnalisaUpdateLSCFibre laid for RFPD audio

 [Annalisa, Jenne, Rana, Steve]

We installed the fibres on cable trays the 1Y2 and the Control Room.

Still to do: find a power supply for the Fiboxes and plug everything in.

  315   Wed Feb 13 20:37:11 2008 JohnUpdateLSCFibre locking - Fiber
Sam and I observed fringes in the light reflected from the Y arm. These fringes are due to the sidebands and not the carrier. To improve matters we plan to reduce the RF AM and increase our modulation index.
  272   Sat Jan 26 02:08:53 2008 JohnOmnistructureLSCFibres
There is now a fibre running from the SP table to the ISCT at the Y-end. In the coming days I will try to mode match the beam from this fibre into the arm through ETMY. To achieve this I will be altering the optical layout of this table.
  296   Mon Feb 4 22:01:57 2008 JohnSummaryLSCFibres auxiliary locking - Fibers
I managed to couple ~75% of the light transmitted from the y arm, through the fibre, back to the SP table. I hoped that this would be a good way to match the beam from the fibre into the arm. Still no flashes. It looks like the cameras just aren't sensitive enough.
  10860   Wed Jan 7 02:54:09 2015 JenneUpdateLSCFiddling with DARM filters

One of the things that we had talked about last night was the totally tiny amount of phase margin that we have in the CARM and DARM loops.  DARM seemed to be the most obnoxious loop last night, so I focused on that today, although the CARM and DARM loops are pretty much identical.

(Q tells me via email that the phase budget has the same ~14 degree discrepancy between what we expect and what we measure as his estimate last night.  However, the Caltech network issues prevented his posting an elog.)

So, we definitely need to figure out where this 14 deg is going, but for now, I wanted to see if I could recover a couple of extra degrees just by modifying the filters.

The original filters do seem to eat a lot of phase:

DARM_design_orig.pdf

The short version of the story is that I didn't leave the filters changed at all.  I reverted back to the last version of the filter file from Monday night, so currently everything is as it was.

I tried increasing the Q of the zeros on the cyan and brown filters, which would sacrifice some gain at ~20 Hz, but hopefully win us 10+ degrees of phase.  This gave me a dip of about a factor of 2 between the new and old filters (all servo filters combined added up to this factor of 2 in magnitude) between ~20Hz - 70Hz. 

When we were locked using DARM for just the Yarm (for the UGF servo commissioning), I took a spectra of the error signal (which was POY) as a reference, then loaded in my new filters.  For the most part, the spectra didn't change (which is good, since the magnitude of the filter didn't change much.).  The spectra was bigger though between 50-70Hz, in kind of a sharp bandpass-looking shape that I wasn't expecting.    I don't know exactly why that's happening.

Anyhow, we tried the new filters once or twice with the full IFO, but kept losing lock.  Since I clearly haven't put in enough thought yet for these (particularly, how much suppression do we really need? what are our requirements???), I reverted back to the filter file from last night.  We continued locking, and checking out the new UGF servo that Diego is elogging about.

  14703   Wed Jun 26 20:45:03 2019 gautamUpdateCamerasField of view options

For the beam spot position tracking, I am wondering if there is any benefit to going for a wider field of view and getting the OSEMs in the frame? It may provide some "anchor points" against which whatever algorithm can calibrate the spot position against. But there are also several point scatterers visible in the current view, and perhaps the Gaussiam beam profile moving over them and tracking the scattered intensity from these point scatterers serves the same function? I don't know of a good solution to have a "switchable" field of view configuration in the already cramped camera enclosure though.

Also, I think it may be useful to have a cron job take a picture of MC2 and archive it (once a week? or daily?) to have some long term diagnostic of how the scattered light received by the camera changes over several months.

Quote:

The GigE is focused now and I have closed the lid. I'm attaching a picture of the MC2 beam spot, captured using GigE at an exposure time of 400µs

  10474   Tue Sep 9 00:34:34 2014 JenneUpdateLSCFiguring out where to do DARM->AS55

This afternoon, after Q and Manasa finished recovering from the activities of the morning, I aligned the IFO, and went to the Yend to touch up the alignment of the green to the arm.  I don't know if it was the alignment (I didn't do the PSL table), or I happened to have a good combination of laser temperatures, or what, but the Yend ALS noise was super good.  After that, the low frequency noise contribution is different lock-to-lock, and I haven't discerned a pattern yet.

One thing that we want to try is to get DARM to AS55 so that we're entirely off of ALS (assuming we've already gotten CARM to sqrtInvTrans).  However, according to Q's simulations, we have to get past arm power of a few before we are within the AS55 linewidth.  I have a DTT running showing me the phase between AS55 and ALSdiff as I reduce the CARM offset, but I haven't been able to get close enough to see the sign flip when CARM is on sqrtInvTrans.  If I just sweep through with both CARM and DARM on ALS, I see the sign flip.  I've tried a few different things, but I have not successfully gotten a transition to AS55 while the arm powers were above 1.  Empirically,  I think I want them at at least 3 or 4.

Koji suggested locking the DRMI rather than PRMI, to widen the AS55 linewidth, but I haven't tried that tonight.  Maybe tomorrow night.

I have made a ruidimentary lockloss plotting script, that I have put in ..../scripts/LSC/LockLossData, but I'm not satisfied with it yet.  Somehow it's not catching the lockloss, even though it's supposed to run when the ALS watch/down scripts run.  I'll need to look into this when I'm not so sleepy.

Q, can you please work on figuring out the phase tracker gain tracker?  It will be nice to have that functional so we don't have to fret about the phase tracker gains. 

Manasa, can you please estimate what kind of mode matching we have on the PSL table between the arm greens and the PSL green?  We *do not* want to touch any optics at this point.  Just stick in a power meter to see how much power we're getting from each beam, and then think about the peak height we see, and what that might tell us about our mode overlap.  If we determine it is total crap, we can think about measuring the beams that go either toward the camera, or the DC PDs, since neither of those paths require careful alignment, and they are already picked off from the main beatnote path.  But first, what is our current efficiency?  Yarm is first, then Xarm, since Yarm seems worse (peak height is larger for non-00 modes!)

  10478   Tue Sep 9 14:25:46 2014 jamieUpdateLSCFiguring out where to do DARM->AS55

Quote:

I have made a ruidimentary lockloss plotting script, that I have put in ..../scripts/LSC/LockLossData, but I'm not satisfied with it yet.  Somehow it's not catching the lockloss, even though it's supposed to run when the ALS watch/down scripts run.  I'll need to look into this when I'm not so sleepy.

We developed a fairly sophisticated lockloss script at the sites, which you could try using as well.  It's at:

USERAPPS/sys/common/scripts/lockloss

It requires a reasonably up-to-date install of cdsutils, and the tconvert utility.  It uses guardian at the sites to determine when locklosses happen, but you can use it without guardian by just feeding it a specific time to plot.  It also accepts a list of channels to plot, one per line.

  5128   Fri Aug 5 20:44:26 2011 jamieMetaphysicsTreasureFilm crew here Monday morning

Just a reminder that a film crew will be here Monday morning, filming Christian Ott for some Discovery channel show.

They are slated to be here from 8am to 12:30pm or so.  They will take a couple of shots inside the lab, and the rest of the filming should be of Christian in the control room (which they will "clean up" and fit with "sexy lighting").  I will try to be here the whole time to oversee everything.

  7092   Mon Aug 6 17:15:15 2012 JenneUpdateASSFilter banks not working

I was trying to load some filters into the ASS so that I can try it out, but for some reason the filter banks aren't working - clicking the on/off buttons doesn't do anything, filters (which exist in the .txt file generated by foton) don't load.

I've emailed cds-announce to see if anyone has any ideas.

  7099   Tue Aug 7 00:22:10 2012 JenneUpdateASSFilter banks working

Quote:

I was trying to load some filters into the ASS so that I can try it out, but for some reason the filter banks aren't working - clicking the on/off buttons doesn't do anything, filters (which exist in the .txt file generated by foton) don't load.

I've emailed cds-announce to see if anyone has any ideas.

 

When the network / fb went bad this afternoon, I had been working on the ASS model, shortening the names of the filter banks to fix the problem from elog 7092.  I wanted to finish working on that, so the ASS model is now rebuilt with slightly shorter names in the filterbanks (which fixes the problem of the filter banks not working).

------------------------------------------------------------

I mentioned this to Jamie the other day, but here's the error that you get when the GoTo/From tags aren't working:

>>rtcds make c1ass
### building c1ass...
Cleaning c1ass...
Done
Parsing the model c1ass...
IPC 9 8 is C1:LSC-ASS_LSC
IPC 9 8 is ISHME
IPC 10 9 is C1:RFM-LSC_TRX
IPC 10 9 is IPCIE
IPC 11 10 is C1:RFM-LSC_TRY
IPC 11 10 is IPCIE

INPUT XARM_LSC_in is NOT connected

INPUT YARM_LSC_in is NOT connected

***ERROR: Found total of ** 2 ** INPUT parts not connected

make[1]: *** [c1ass] Error 255
make: *** [c1ass] Error 1

I don't know why these tags weren't working, but there was a GoTo tag on the output of the LSC shmem block, and then Froms on each of the XARM_LSC_in and YARM_LSC_in.  The other day I played around with a bunch of different things (grounding inputs, terminating outputs, whatever), but finally replacing the tags with identical ones freshly taken from CDS_PARTS made it happy. 

  5730   Mon Oct 24 19:48:16 2011 MirkoUpdateAdaptive FilteringFilter execution time

Toyed around some more with the adaptive filters.

Execution time:

nTaps    Downsampling factor     Execution time average / max in ca. 3 min [us], (480 us available)
1000     16                                110 / 150
2000     16                                280 / 340
3000     16                                380 / 470
4000     16                                Over limit

Now we are running with Downsampling 32, 4000 Taps => max 410us execution time.

I tried to desynchronize the downsampled operations of the filters of the different DOFs. That however increased execution time by about 10%. So I undid that.

 

  3802   Thu Oct 28 02:01:51 2010 KevinUpdatePSLFilter for 2W Laser

[Rana and Kevin]

I made a low pass filter for the piezo driver for the 2W laser that is now installed. The filter has a pole at 2.9 Hz. The transfer function is shown in attachment 1.

Attachment 2 shows the outside of the filter with the circuit diagram and attachment 2 shows the inside of the filter.

  3707   Wed Oct 13 17:12:33 2010 josephb, yutaUpdateCDSFilter name length problem found and fixed

The missing filter files for ULPOS, URPOS, and so forth for the mode cleaner optics was due to the length of the names of the filters. 

This was not a problem for the c1sus model because it was using its own name as the first 3 letters of its designation.  A filter for the sus model would be called something like BS_TO_COIL_MTRX_0_0, while for the mcs it would be called SUS_MC1_TO_COIL_MTRX_0_0, an extra 4 characters.

However, the c1mcs model used the "top_name" feature which uses a subsystem box within the simlink model to rename all the channels.  Apparently in the filter file, this means it has to add the top name to the front of everything, adding an additional 3 characters.  This pushed things over the length limit.

A hard cap of 18 characters has been added to the FiltMuxMatrix.pm file (located in /cvs/cds/caltech/c1/, so that it will prevent this type of problem in the future by stopping at compile time and presenting a helpful error message.

I also fixed a bug with too many spaces in the feCodeGen.pl file when dealing with top_names and the filtMuxMatrix.pm preventing some .adl files from being generated.

Also of interest, MC3 appears to never have had F2A filters.  For the moment we're running without them, but since they're just a fine tuning it shouldn't affect locking tonight.

 

Improbability factor of mode cleaner suspensions working tonight: ~20

 

  995   Fri Sep 26 00:19:54 2008 JenneUpdatePSLFilter-action with the PMC
Written, but not posted on 24Sept2008:



PMC adventures for this evening



Today's mission was to make more progress on increasing the bandwidth of the PMC servo.



First order of business was to improve the performance of the 14.6kHz notch that Rana put in the PMC servo board a few weeks ago to remove the 14.6kHz body mode resonance of the PMC. Looking at the zoomed in TF that I posted Monday (elog #978), we see that there is still a remnant of a peak near 14.5kHz. A first gut-reaction is that the notch is not tuned properly, that we have just missed the peak. As previously noted in the elog, the peak that we are trying to notch out is at 14.68kHz (elog #874). By unlocking the PMC and measuring the transfer function between FP2 and OutMon (OutMon is the monitor for the high voltage going to the PMC's PZT), I measure the transfer function of the notch, and find that it is notching at 14.63kHz. So we're a teensy bit off, but the Q of the notch is such that we're still getting improvement at the peak frequency. After checking that we are hitting the correct frequency, I put a short (just some wire) around R21, which is the R in the RLC notch filter, to increase the depth of the notch. At the peak frequency of 14.68kHz, we see a 2.5dB improvement of the notch. At the actual notch frequency of 14.63kHz, we see a 3.2dB increase in the depth of the notch. So, shorting R21 helped a little, but not a lot. Also, it's clear that we don't get that much more improvement by being on the resonant frequency, so there's no need to go in and tune the notch on the board.



Second order of business was to investigate the 18.34kHz peak in the transfer function. (Rana spent some time Monday night measuring this peak, and determined that it was at 18.34kHz) We decided that the best plan was to re-implement the Pomona Box notch filter that had previously existed to remove a higher frequency body mode, but tuned for the 18.34kHz mode. I am still not entirely sure what this mode is, but clearly it's a problem by about 20dB (on the TF, the next highest peak is 20dB below the 18.34kHz peak). Unfortunately, while the components should, by Matlab calculations, give me an 18.3kHz notch, I ended up with something like a 21.7kHz notch. This notch is approximately -30dB at 21.7kHz, and -20dB at 18.3kHz. I still need to take transfer functions and power spectra of the PMC servo with this new filter in place to (a) confirm that it did some good, and (b) to determine how important it is that the notch be right-on. More likely than not, I'll take the filter out and fiddle with the capacitors until I get the correct notch frequency.



Third on the list was to lock everything back up (FSS, PMC) after my tinkering, and see what kind of gain we get. Rob and I fiddled with the PMC gain, and it looks like the servo oscillates just before we get up to the max slider gain of 30dB. Looking at the power spectra in DTT, we do not see any significant peaks that suggest oscillation, so it is likely that there is some investigation to be done at frequencies above the 7kHz that we were able to look at with DTT (which isn't surprising, since all of this work has been at 14kHz and higher).



A final note is that we see a feature around 9kHz in the transfer function, and it is not at all clear where it comes from. At this time, it does not seem to be the dominant feature preventing us from increasing the gain, but at some point if we want the bandwidth of the PMC servo to be 10kHz, we'll have to figure this one out.



Still on the PMC todo list:


  • Measure the new transfer function, see if 18.34kHz peak is reduced

  • Tune Pomona Box notch filter to 18.3kHz instead of the current 21.7kHz

  • Retake power spectra of different items on top of PMC, compare to see if there is any one configuration that it obviously better than the others.

  • Find out why the PMC still oscillates when we try to take it up to the max slider gain, and fix it.

PS, is anyone else having trouble getting to the elog from laptops on other parts of the Caltech network (but not LIGO network)? My laptop won't go to the elog, but I can get to the rest of the internet using the Caltech wireless. My computer stopped seeing the elog on Tuesday or so. Joe, do you have any inspiration? Thanks.
  522   Fri Jun 6 11:19:13 2008 CarynSummaryPEMFiltering MC_L and MC_F with PEM:ACC and microphone
Tried to filter MC_L and MC_F with acc/seis data and microphone data using wiener filter (levinson)

-Used get_mic_data.m and miso_filter_lev.m to make SISO filter for 2 minutes of IOO-MC_F data. Used PEM-AS_MIC signal as noise input data. Filters calculated at initial time were applied to later data in 1 hour intervals.
-microphone filter did not seem to filter MC_F very well in high frequency range using this filtering procedure.
-residual is larger than est (see MC_F pdf)
-Used do_all_time_lev.m to make graph of max(rms(residual)) to N(order) for different times.(note for each N, filter was calculated for initial time and then applied to data at other times).
-relation of max(rms(residual)) to N(order) is time sensitive (note-on graph, time interval is 1hour) (see MC_F pdf)
-Presumably, max(rms(residual)) should decrease as N increases and increase as time increases since the filter probably becomes worse with time. I think the reason this isn't always true in this case is that the max(rms(residual)) corresponds to a peak (possibly a 60Hz multiple) and the wiener filter isn't filtering out that peak very well.


-Used get_z_data.m and miso_filter_lev.m to make MISO filter for 2 minutes of IOO-MC_L used the following signals as noise input data
PEM-ACC_MC1_X
PEM-ACC_MC2_X
PEM-ACC_MC1_Y
PEM-ACC_MC2_Y
PEM-ACC_MC1_Z
PEM-ACC_MC2_Z
PEM-SEIS_MC1_Y
-Filter was applied to later data in 2hour intervals.
-Used do_all_time_lev.m to make graph of max(rms(residual)) to N(order) for different times.(note for each N, filter was calculated for initial time and then applied to data at other times).
-acc/seis filter seemed to filter MC_L OK for 128,256,512Hz srates. 64 Hz wasn't ok for certain N's after a period of time.
-residual is smaller than est for srates not 64Hz (see MC_L pdf)
-residual is larger than est for 64Hz at N=1448 for later times (see MC_L pdf)
-relation of max(rms(residual)) to N is not as time sensitive for higher sample rates (note-on graph, time interval is 2hours) (see MC_L pdf). Perhaps the levinson 64Hz sample rate filter doesn't do as well as time passes for these signals. When the filter didn't do well, the max(rms(residual)) seemed to increase with N.
-For 512Hz sample rate filter the max(rms(residual)) decreased with time. If the max(rms(residual)) were an indication of filter performance, it would mean that the 512Hz filter calculated at the initial time was performing better later as hours passed by! Perhaps max(rms(residual)) isn't always great at indicating filter performance.

Programming notes
-I had to modify values in do_all_time_lev.m to get the program to loop over the srates,N's,times I wanted
-do_all_time_lev.m is not as clean as do_all_lev.m
-for making the plots do_all_lev.m (which isn't really a procedure and is messy) has some examples of how to plot things from do_all_time_lev.m.
  6049   Wed Nov 30 02:04:26 2011 rana, den, jenne, kiwamu, jzweizigUpdateCDSFiltering Noise issue tracked down ???

 You can read through all of our past tests to see what didn't work in tracking things down. As Den mentions, there was actually a lot of evidence that there was some double->single precision action in the filter calculation causing the noise we saw.

However, it turns out that this is NOT the case.

This afternoon I was so confused that I enlisted JZ to help us out. He came over and I tried to replicate the error. When looking at the time series, we noticed that it wasn't random noise; the signals seem to be getting clipped as they crossed zero. Sort of like a stiction problem. JZ left to go replicate the error on an offline system.

This turned out to be the important clue. As we examine the code we find this inside of fm10Gen.c:

if((new_hist < 1e-20) && (new_hist > -1e-20)) new_hist = new_hist<0 ? -1e-20: 1e-20;

this is line is basically trapping the filter history at 1e-20, to prevent some kind of numerical underflow problem (?). Seems reasonable, except that some filters which have higher order low passing in them actually have an overall scale factor which can be small (even as small as 1e-23, as Den pointed out).

So the reason we saw such weird behavior is that the first filter in SUSPOS is an AC coupling filter. This takes the OSEM signal and remove the large mean value. Then the next filter multiplies it by 1e-23 before doing the filtering and you end up with this noise in the filter history.

I looked and this line is commented out in the new BiQuad code, but as far as I can tell this issue has been around in aLIGO, eLIGO, iLIGO, etc. for a long time and could have been causing many cases of excess noise whenever we ended up a tiny gain factor in an IIR filter. At the 40m, there are easily a hundred such cases.

For now, I suppose we can just change this number to 1e-40 or so. I don't know how to calculate what the right number should be. Not sure why this underflow is not an issue for the BiQuad, however.

  6051   Wed Nov 30 11:04:26 2011 josephbUpdateCDSFiltering Noise issue tracked down ???

Quote:

For now, I suppose we can just change this number to 1e-40 or so. I don't know how to calculate what the right number should be. Not sure why this underflow is not an issue for the BiQuad, however.

According to the RCG SVN logs, the reason it was removed was a more general change done to the compiled code, not specific to just the biquad.  Basically, the ability to have an underflow number (subnormal) has been turned off completely by having any number that underflows set to zero. I'm not positive, but from a quick search looks that the smallest number before hitting is an underflow as a double is 2.2250738585072014e-308.

Alex's entry from the SVN log for 2663:

Added new fz_daz() function to turn on two bits in the FPU SSE control register.
Bits FZ (flush underflows to zero) and DOZ (denorms are zeros) are set to
avoid runaway code on float/double denorms (really small numbers).
Ref: http://software.intel.com/en-us/articles/how-to-avoid-performance-penalties-for-gradual-underflow-behavior/

SVN log 2664:

Removed +- 1e-20 limiting code, this is taken care of by setting FZ/DOZ bits
in the CPU SEE control register (see mathInline.h)

SVN log 2665:

Kill the underflows and roll down float denorms to zero,
see fz_doz() in mathInline.h.

  6052   Wed Nov 30 11:36:12 2011 DenUpdateCDSFiltering Noise issue tracked down ???

Quote:

 if((new_hist < 1e-20) && (new_hist > -1e-20)) new_hist = new_hist<0 ? -1e-20: 1e-20;

20 is indeed a random number. We can change it to 300. Alex said that during that iir filter calculations sometimes numbers are very small and if they are less then 1e-308 then a very slow code in the processor is executed and this will crash the online system. For single precision this number is 1e-38 and may be 10 years ago it was not decided for sure what to use - float or double. 20 will be "OK" for both but as we can see causes other problems.

Anyway, Alex removed this line from the code and added another code that sets the two proper bits in the MXCSR register and prohibits to the CPU to run the slow code. As far as I understand if the numbers are less then 1e-308 they become 0. Roughly, this is equivalent to 

 if((new_hist < 1e-308) && (new_hist > -1e-308)) new_hist = 0;

This is in 2.4 release. It is in the svn. I think we can install it and figure out if the problem is gone.

 

  7074   Wed Aug 1 22:37:21 2012 JenneUpdateASSFilters installed in the ASS

As part of trying to figure out what is going on with the ASS, I wanted to figure out what filters are installed on which lockins.

Each "DoF"(1-6) has a zpk(0.1,0.0001,1)gain(1000), which is a lowpass with 60dB of gain at DC, and unity gain at high frequencies.

For the lockins, since there are so many, I made a spreadsheet to keep track of them (attached). 

So, what's the point?  The point is, I think that all of the LOCKIN_I filter modules should be the same, with a single low pass filter.  The Q filter banks don't matter, since we don't use those signals, and the signals are grounded inside the model.  The phase of each lockin was / should be tuned such that all of the interesting signal goes to I, and nothing goes to Q.  The SIG filter modules seem okay, in that they're all the same, except for their band pass frequency.  I just need to check to see what frequency the ASS scripts are trying to actuate at, to make sure we're bandpassing the correct things.

 

  11590   Thu Sep 10 09:37:34 2015 IgnacioSummaryIOOFilters left on MCL static module

The following MCL filters were left loaded in the T240-X and T240-Y FF filter modules (filters go in pairs, both on):

FM7: SISO filters for MCL elog:11541

FM8: MISO v1 elog:11547

FM9: MISO v1.1 Small improvement over MISO v1

FM10 MISO v2 elog:11563

FM5 MISO v3.1 elog:11584 (best one)

FM6 MISO 3.1.1 elog:11584 (second best one)

 

  10239   Fri Jul 18 19:32:50 2014 AkhilSummaryElectronicsFilters used inside the Frequency Counter

 

 Thanks Koji , for your  hint for the brain teasing puzzle. I was looking into Filters that are usually used in devices like counters, DSO and other scopes. I found that , to improve the quality of the measurement one of the best approach  is averaging. I looked deeper into averaging and found out this:

There are two general use-cases for averaging . The first, successive sample averaging, takes a single acquisition and averages between its samples. The second, successive capture averaging, combines the corresponding  samples of multiple captures to create a single capture. Successive sample averaging is also called boxcar filtering or moving average filtering. In an implementation of this type of averaging each output sample represents the average value of M consecutive input samples. This type of averaging removes noise (one of the reasons the noise level was not bad: http://nodus.ligo.caltech.edu:8080/40m/10151) by decreasing the device's bandwidth(could be one of the reasons why the FC operates in 4 different frequency ranges). It applies an LPF function with a 3dB point approximated by  0.433 * s / M, where M is the number of samples to be averaged, and s is the sample rate in samples per second. 

Now I tried verifying the 3 dB points in the gain plots I generated :

For 1 s Sampling time : the 3 dB point for such a Boxcar filter should be at 0.433* 1/M. If we assume that it averages for 2 samples, M=2 which gives the 3dB point at 0.288 Hz but occurs somewhere between 0.3 and 0.4 Hz.  (http://nodus.ligo.caltech.edu:8080/40m/140619_120548/GainVsFreq.png)

For 0.1s Sampling time: the 3dB point should be at 2.17 Hz and in reality is 2.5 Hz(http://nodus.ligo.caltech.edu:8080/40m/140701_211904/gain.png).

Also, This type of filter will have very sharp nulls at frequencies corresponding to signals whose periods are integer sub-multiples of M/s. As seen my previous plots (http://nodus.ligo.caltech.edu:8080/40m/10118 , http://nodus.ligo.caltech.edu:8080/40m/10070) there are sharp nulls at frequencies

0.4 Hz for 1S sampling time and

at 1.5 Hz,3 Hz for 0.1 S sampling time as correctly predicted.

The moving average filter is  L-sample moving average FIR, with the frequency response as:   H(ω) = (1/L) (1 − e− jω L)/(1 − e− jω)..

There is an overall delay of (M - 1)/2 samples from such a length-M causal FIR filter. 

The expected bode plots for such a filter with L= 5 is attached(attachment 2).

  10246   Mon Jul 21 12:16:27 2014 AkhilSummaryElectronicsFilters used inside the Frequency Counter

The expected bode plots for such a filter with L= 4 is attached and compared with the measured.

RXA: When comparing two things, please put them onto the same plot so that they can be compared.

  3437   Wed Aug 18 19:19:38 2010 JenneUpdateSUSFinal 2 TTs suspended!

[Jenne, Yoichi]

The final 2 Tip Tilts (#1 and #5) have been suspended.  We have designated #5 the spare.  It looks like there might be a teensy bit of dust on the AR surface of the optic in #5, right near the edge of the coating.  Not a critical issue if this one is the spare, although we should see if we can blow it off with the Nitrogen.  Both #1 and #5's optics were suspended using the thicker wire, 0.0036" diameter.  This leaves 4/5 TTs with this thick wire, and 1 of the 5 has the thin wire.

To do still: Balance both #1 and #5, and then measure the modes of each.  Then we'll be ready to install them into the chambers, and we'll reserve #5 for shake table TFs for some later date.

 

  8113   Wed Feb 20 01:40:37 2013 ManasaUpdateAlignmentFinal IFO alignment- in progress

[Yuta, Sendhil, Jamie, Jenne, Rana]

1. After the MC centering, we tried to align the IFO using IPPOS and IPANG as reference. This did not recover the alignment perfectly. We were clipping at the BS aperture. Using TTs, we centered the beam at BS and PRM.
2. Using TTs, the beam was centered at ITMY and ETMY.
3. IPPOS and IPANG mirrors in-vacuum were aligned and were centered at the out-of-vacuum optics.
4. We checked the centering of the beam on optics in the BS and ITMY chamber. (Yuta will make an elog with the layout)
5. We retro-reflected ITMY at the BS and aligned ETMY such that we saw a couple of bounces in the arm cavity.
6. Using BS, the beam was steered to go through the center of ITMX and ETMX.
7. At this point we were able to see the MI fringes at the AS port.
8. We made fine alignments to the ITMX such that we saw MI reflected at the Farday.
9. We retro-reflected ITMX and aligned ETMX to see the beam bounce at the ITMX.
10. We aligned PRM such that PRC flashes. But we were not happy with the flashes (they were in higher order modes). We suspect that minor tuning of the input pointing might be necessary.
11. We closed for the day

  8338   Mon Mar 25 16:51:43 2013 ChloeUpdate Final QPD Circuit Design

This is the final version of the QPD circuit I'm going to build. After playing around with the spatial arrangement, this should fit into the box that I was planning to use, although it will be a rather tight fit. The pitch, yaw, and summing circuit will be handled with a quad op amp. Planning to meet with Eric tomorrow to figure out the logistics of building things.

In the meantime, I'm reading about designing the ECDL for my summer project with Tara. He sent me several papers to read so we can talk on Wednesday.

  11535   Fri Aug 28 00:59:55 2015 IgnacioUpdateIOOFinal SISO FF Wiener Filter for MCL

This is my final SISO Wiener filter for MCL that uses the T240-X seismo as its witness.

The main difference between this filter and the one on elog:11532 is the actual 1/f rolloff this filter pocesses. My last filter had a pair of complex zeroes at 2kHz, that gave the filter some unusual behavior at high frequencies, thanks Vectfit. This filter has 10 poles and 8 zeroes, something Vectfit doesn't allow for and needs to be done manually.

The nice thing about this filter is the fact that Eric and I turned this filter on during his 40 min PRFPMI lock last night, Spectra for this is coming soon.

This filter lives on the static Wiener path on the OAF machine, MCL to MC2, filter bank 7.

Anyways, the usual plots are shown below. 

 

Filter:

T240-X (SISO)

 

Training data + Predicted FIR and IIR subtraction:

 

Online subtraction results:(High freq. stuff shown for noise injection evaluation of the filter)

MCL

YARM

Subtraction performace:

  14764   Tue Jul 16 15:17:57 2019 KojiHowToCDSFinal bit bug of the BIO CDS module

Yutaro talked about the BIO bug in KAGRA elog. http://klog.icrr.u-tokyo.ac.jp/osl/?r=9536

I think I made the similar change for the 40m model somewhere (don't remember), but be aware of the presense of this bug.

  14890   Tue Sep 17 14:43:59 2019 gautamHowToCDSFinal bit bug of the BIO CDS module

Came across this while looking up the BIO situation at 1Y2. For reference, the fix Koji mentions can be seen in the attached screenshot (one example, the other BIO cards also have a similar fix). The 16th bit of the BIO is grounded, and some bit-shifting magic is used to implement the desired output.

Quote:

Yutaro talked about the BIO bug in KAGRA elog. http://klog.icrr.u-tokyo.ac.jp/osl/?r=9536

I think I made the similar change for the 40m model somewhere (don't remember), but be aware of the presense of this bug.

  14941   Fri Oct 4 22:22:03 2019 gautamUpdateCDSFinal incarnation of latch.py

[KA, GV]

This elog is meant to be a summary of some of the many subtleties on the CM board. The latest schematic of the version used at the 40m can be found at D1500308 .

Latch logic:

  • There are several Binary Outputs and one Binary Input to the CM board.
  • The outputs control ENABLE/DISABLE switches and gains of amplifier stages, while the input reports whenever the limiter has been reached.
  • The variable gain feature is implemented by enabling/bypassing several cascaded fixed gain stages. So in order to change the gain of a single composite amplifier stage, multiple individual amplifier stages have to be switched.
  • This is implemented by the user interacting with the hardware via a "control word", consisting of a number of bits depending on the number of cascaded stages that have to be switched. 
  • This control word is sent to the device via modbus EPICS, which is an asynchronous communication protocol. Hence, it may be that the individual bits composing the control word get switched asynchronously. This would be disastrous, as there can be transient glitches in the gain of the stage being controlled. 
  • To protect against such problems, there is a latch IC in the hardware between the Binary Inputs to the board (= Binary Outputs from Acromags), and the actual switches (= MAX333) that enable/bypass the cascaded gain stages. The latch IC used is a SN74ALS573. This device acts as a bus, which transmits/blocks changes for multiple bits (= our control word) from propagating, depending on the state of a single bit (= the LATCH ENABLE bit). Thus, by controlling a single bit, we can guarantee that multiple bits get switched synchronously
  • In order to use this latch capability, we need some software logic that sets/disables the LATCH ENABLE bit. For our system, this logic is implemented in the form of a continuously running python 🐍 script, located at /cvs/cds/caltech/target/c1iscaux/latch.py. It is implemented as a systemctl service on the c1iscaux Supermicro. The logic implemented in this script is shown in Attachment #1. While the channels referred to in that attachment are for REFL1_GAIN, the same logic is implemented for REFL2_GAIN, AO_GAIN, and the SuperBoosts.
  • Some FAQ:
    1. Q: Why do we need the soft channels C1:LSC-REFL1_SET_LSB and C1:LSC-REFL1_SET_MSB?
      A: These soft channels are what is physically linked to the Acromag Binary Outputs. In order for our latch logic to be effective, we need to detect when the user asks for a change, and then disable the LATCH ENABLE bit (which is on by default, see FAQ #3) before changing the physical acromag channels. The soft channels form the protective layer between the user and the hardware, allowing latch.py to function.
    2. Q: Why is there an "_MSB" and "_LSB" soft channel? 
      A: This has to do with the mbboDirect EPICS channel type, which is used to control the multiple bits in our control word using a single input (= an MEDM gain slider). The mbboDirect data-type requires the bits it controls to have consecutive hardware addresses. However, the Acromag hardware addressing scheme is not always compatible with this requirement (see pg 33 of the manual for why this is the case). Hence, we have to artifically break up the control word into two separate control words compatible with the Acromag addressing scheme. This functionality is implemented in latch.py.
    3. Q: Why is the default state of LATCH ENABLE set to ON? 
      A: This has to do with the fact that all Binary Inputs, not just the multi-bit ones, to the CM board are propagated to the control hardware via a latch IC. For the single-bit channels, there is no requirement that the switching be synchronous. Hence, rather than setting up ~10 more single-bit soft channels and detecting changes before propagating them, we decided to leave the LATCH ENABLE ON by default, and only disable it when changing the multi-bit gain channels. This is the same way the logic was implemented in the VME state code, and we think that there are no logic reasons why it would fail. But if someone comes up with something, we can change the logic.

Acromag BIO testing:

During my bench testing of the Acromag chassis, I had not yet figured out mbboDirect and the latch logic, so I did not fully verify the channel mapping (= wiring inside the Acromag box), and whether the sitching behavior was consistent with what we expect. Koji and I verified (using the LED tester breakout board) that all the channels have the expected behavior 👏. Note that this is only a certification at the front-panel DB37 connectors of the Acromag chassis  testing of the integrated electronics chain including the CM board is in progress...

ELOG V3.1.3-