  1665   Wed Jun 10 09:06:12 2009 steveUpdatePEMparticle counts and turbulance

I moved the mobile HEPA filter from ITMX's north door to ITMX-ISCT and covered it up with a merostate tent to accommodate the aluminum foil particle measurement on June 5

It lowered the 40m baseline counts by about a factor of 3 of 0.5 micron and a factor of 2 of 1.0 micron.

The HEPA filter is sweeping the floor and blowing the particles upwards. The MET ONE counter is on the top of the IOOC looking south at ~75 degrees upward.

 

Attachment 1: parturbulnc.jpg
parturbulnc.jpg
  1666   Wed Jun 10 09:28:14 2009 steveUpdatePSLPMC

The PMC alarm was on this morning. It was relocked at lower HV

The FSS_RMTEMP jumped 0.5 C so The PZT was compensating for it.

 

Attachment 1: pmctemphv.jpg
pmctemphv.jpg
  1667   Thu Jun 11 03:15:15 2009 AlbertoUpdateLSCDD Handoff attempts

Pete, Alberto,

tonight we worked on the tuning of the double demod phases for the handoff of the short DOFs control signals.

Only MICH can now undergo the handoff. PRC can't make it.

Basically, we tuned the PD6 demod phase and reduced the offset in PD6_I. Then we tuned the relative gain of PD6_I and PD2_I so that the two open loop transfer function of the control loops would match. We tried that in several ways and several times but without success.

I guess we're missing to do/check something.

  1668   Thu Jun 11 14:54:18 2009 josephb, albertoUpdateComputersWireless network

After poking around for a few minutes several facts became clear:

1) At least one GPIB interface has a hard ethernet connection (and does not currently go through the wireless).

2) The wireless on the laptop works fine, since it can connect to the router.

3) The rest of the martian network cannot talk to the router.

This led to me replugging the ethernet cord back into the wireless router, which at some point in the past had been unplugged.  The computers now seem to be happy and can talk to each other.

 

  1669   Thu Jun 11 22:14:10 2009 AlbertoUpdateLSCDD Handoff for the Short DOFs completed

This afternoon I tuned the handoff script for the SRC, after that Rob eralier during the day had already adjusted that for PRC. To do that, I followed the procedure in the Wiki.

  • I measured the OL transfer function of the single demod path and of the double demod path and tuned thier gains so that they matched
  • I tuned the double demod pahses of PD_7 and PD_8 in order to reduce the offset in the PD_x_I signals

After that the SRC could get locked with the double demod signals. the open loop transfer function emasurement on the PRC loop showed that it was nearly unstable. Rob reduced a little its gain to improve the stability.

The DD handoff is now working and we can get back to locking the interferometer.

  1670   Fri Jun 12 02:01:03 2009 robUpdateComputer Scripts / ProgramsDRM matrix diagonalization

I started two scripts, senseDRM and loadDRMImatrixData.m, which Peter will bang on until they're correct.  They're in the $SCRIPTS/LSC directory.  The first is a perl script which uses TDS tools to drive the DRM optics and measure the response at the double demod photo-detectors, and write these results to a series of files loadable by matlab.  The second loads the output from the first script, inverts the resulting sensing matrix to get an input matrix, and spits out a tdswrite command which can be copied and pasted into a terminal to load the new input matrix values. 

What's left is mainly in figuring out how to do the matrix inversion properly.  Right now the script does not account for the output matrix, the gains in the feedback filters at the measurement frequency, or the fact that we'll likely want the UGF of our loops to be less than the measurement frequency.  Peter's going to hash out these details.

  1671   Fri Jun 12 09:56:38 2009 steveConfigurationVACnew maglev installed

Joe and Steve

 

The retrofitted Osaka 390 was installed on the pumpspool yesterday.

V1 gate valve is disabled for safety by disconnected pneumatic power plug.

The foreline of this maglev now have a KF25 size viton o-ring directly on the turbo.

This is bad for leak hunting.

Joe is ready with new interface cable. Power supply and cables are in place.

The maglev was pumped down this morning.

All new gas kits and metal hose were leak checked by sprayed methanol.

There is no obvious sign of leak. I was expecting the pressure to drop below 1e-5 Torr in one hour.

TP2 is drying out the levitating coils of the turbo at ~7 l/s for N2

We'll start the pump as soon as Joe is in.

 

 

Attachment 1: m390pd.jpg
m390pd.jpg
  1672   Fri Jun 12 15:34:49 2009 steveConfigurationVACmaglev is running

Quote:

Joe and Steve

 

The retrofitted Osaka 390 was installed on the pumpspool yesterday.

V1 gate valve is disabled for safety by disconnected pneumatic power plug.

The foreline of this maglev now have a KF25 size viton o-ring directly on the turbo.

This is bad for leak hunting.

Joe is ready with new interface cable. Power supply and cables are in place.

The maglev was pumped down this morning.

All new gas kits and metal hose were leak checked by sprayed methanol.

There is no obvious sign of leak. I was expecting the pressure to drop below 1e-5 Torr in one hour.

TP2 is drying out the levitating coils of the turbo at ~7 l/s for N2

We'll start the pump as soon as Joe is in.

 

 

 Joe and Steve

 

The Maglev is running at 680 Hz, 40,800 RPM with V1 gate valve  closed and  valve disabled to change position. C1vac2 was rebooted before starting.

Interlocks are not tested yet, but the medm COVAC_MONITOR.adl screen is reading correctly. RGA scan will determine the need for baking on Monday

The foreline pressure is still  ~2e-5 Torr

Acceleration takes 3 minutes 30sesconds without load. There is no observabale temp effect on the body of the turbo during braking and acceleration.

 

The IFO is still pumped by the CRYO only

  1673   Mon Jun 15 15:17:33 2009 josephb, SteveConfigurationVACVacuum control and monitor screens

We updated the vacuum control and monitor screens  (C0VAC_MONITOR.adl and C0VAC_CONTROL.adl).  We also updated the /cvs/cds/caltech/target/c1vac1/Vac.db file.

1) We changed the C1:Vac-TP1_lev channel to C1:Vac-TP1_ala channel, since it now is an alarm readback on the new turbo pump rather than an indication of levitation.  The logic on printing the "X" was changed from X is printed on a 1 = ok status) to X is printed on a 0 = problem status.  All references within the Vac.db file to C1:Vac-TP1_lev were changed.  The medm screens also now are labeled Alarm, instead of Levitating.

2) We changed the text displayed by the CP1 channel (C1:Vac-CP1_mon in Vac.db) from "On" and "Off" to "Cold - On" and "Warm - OFF".

3) We restarted the c1vac1 front end as well as the framebuilder after these changes.

  1674   Mon Jun 15 16:31:36 2009 steveConfigurationVACRGA is scanning new Maglev

Quote:

Quote:

Joe and Steve

 

The retrofitted Osaka 390 was installed on the pumpspool yesterday.

V1 gate valve is disabled for safety by disconnected pneumatic power plug.

The foreline of this maglev now have a KF25 size viton o-ring directly on the turbo.

This is bad for leak hunting.

Joe is ready with new interface cable. Power supply and cables are in place.

The maglev was pumped down this morning.

All new gas kits and metal hose were leak checked by sprayed methanol.

There is no obvious sign of leak. I was expecting the pressure to drop below 1e-5 Torr in one hour.

TP2 is drying out the levitating coils of the turbo at ~7 l/s for N2

We'll start the pump as soon as Joe is in.

 

 

 Joe and Steve

 

The Maglev is running at 680 Hz, 40,800 RPM with V1 gate valve  closed and  valve disabled to change position. C1vac2 was rebooted before starting.

Interlocks are not tested yet, but the medm COVAC_MONITOR.adl screen is reading correctly. RGA scan will determine the need for baking on Monday

The foreline pressure is still  ~2e-5 Torr

Acceleration takes 3 minutes 30sesconds without load. There is no observabale temp effect on the body of the turbo during braking and acceleration.

 

The IFO is still pumped by the CRYO only

 The new Maglev fore line pressure is at 4e-6 torr at day 3

Valve VM1 was closed to isolate IFO from RGA and valve VM2 was opened so the RGA can scan the Maglev only.

 

  1675   Tue Jun 16 02:09:31 2009 robUpdateLockingDD handoff finally working

I had trouble getting the SRC handoff from SD to DD to work.  I tried different gains, flipping the PD7 & 8 demod phases by 180 degrees, and messing with the output matrix to reduce cross-couplings in the state with MICH & PRC on DD and SRC on SD.  Eventually I decided to try to make the DRM matrix diagonalization work. 

It does, mostly.  The handoff is now stable, and the loops all have UGFs around 100Hz.  So, tonight anyways, it's possible to run senseDRM and then loadDRMImatrixData.m and run the resulting tdswrite command, and have a working handoff.  I had to eliminate a few PDs (PD5 & PD10) to get it to work properly. 

It would be nice if this script would measure all the PDs and allow individual setting of loop UGFs and measurement frequencies. 

 

 

  1676   Tue Jun 16 03:43:04 2009 robUpdateLockingsame troubles

Lock is still being lost, right at the end of the process when trying to reduce the CARM offset to zero.

  1677   Tue Jun 16 09:02:30 2009 steveConfigurationVACnew maglev RGA scan

The maglev fore line pressure is 3e-6 torr at CC2 after 4 days of pumping. Varian turbo V-70 is pumping on it through V4 and VM2

Actual pumping speed ~10 l/s for N2. There was no baking. The maglev performance looks good : 3e-9 torr on CC4  with RGA -region only.

Attachment 1: nmagd4scan.jpg
nmagd4scan.jpg
Attachment 2: nmagd4.jpg
nmagd4.jpg
  1678   Tue Jun 16 14:02:16 2009 robUpdateLockinglocked

The IFO is locked, at the operating point (zero CARM offset).  The problem with reducing the residual CARM offset in the last stage appears to have been because the common mode gain was getting too high, and so the loop was going unstable at high frequencies. 

The cm_step script is currently a confusing mess, with all the debugging statements.  I'll clean it up this afternoon and check that it still works.

  1679   Tue Jun 16 16:10:01 2009 peteUpdateLockinginput matrix experiments

Last night Rob ran senseDRM and loadDRMImatrixData and came up with the following for the input matrix:

tdswrite C1:LSC-ITMTRX_b2 0.065778 \
C1:LSC-ITMTRX_d2 2.2709 \
C1:LSC-ITMTRX_f2 2.9361 \
C1:LSC-ITMTRX_122 0.42826 \
C1:LSC-ITMTRX_b3 -0.064839 \
C1:LSC-ITMTRX_d3 -0.016913 \
C1:LSC-ITMTRX_f3 -0.021576 \
C1:LSC-ITMTRX_123 -0.0025243 \
C1:LSC-ITMTRX_b5 0.3719 \
C1:LSC-ITMTRX_d5 1.3109 \
C1:LSC-ITMTRX_f5 -0.16412 \
C1:LSC-ITMTRX_125 0.39574 \
C1:LSC-ITMTRX_33 0 \
C1:LSC-ITMTRX_42 0 \
C1:LSC-ITMTRX_155 0

Today, I reran these and got the following, and DD_handoff remained happy:

tdswrite C1:LSC-ITMTRX_b2 -0.10329 \
C1:LSC-ITMTRX_d2 2.0344 \
C1:LSC-ITMTRX_f2 3.2804 \
C1:LSC-ITMTRX_122 0.22516 \
C1:LSC-ITMTRX_b3 -0.076292 \
C1:LSC-ITMTRX_d3 -0.014603 \
C1:LSC-ITMTRX_f3 -0.12101 \
C1:LSC-ITMTRX_123 0.0054128 \
C1:LSC-ITMTRX_b5 0.33521 \
C1:LSC-ITMTRX_d5 1.1425 \
C1:LSC-ITMTRX_f5 -0.32759 \
C1:LSC-ITMTRX_125 0.25877 \
C1:LSC-ITMTRX_33 0 \
C1:LSC-ITMTRX_42 0 \
C1:LSC-ITMTRX_155 0

I wanted to remeasure with the canonical output matrix (-0.7 from MICH to PRM and 0.7 from MICH to SRM), but the DRM freaked out when MICH to PRM went below -0.3.

  1680   Tue Jun 16 17:38:35 2009 robUpdateLockingDeWhites ON

With the common mode servo bandwidth above 30kHz and the BOOST on (1), I was able to switch on the test mass dewhitening.  Finally.

  1681   Tue Jun 16 20:03:41 2009 AlbertoUpdateElectronicsRequirements on Wenzel Multiplier

For the 40m Upgrade, we plan to eliminate the Mach-Zehnder and replace it with a single EOM driven by all three modulation frequencies that we'll need: f1=11MHz, f2=5*f1=55MHz, fmc=29.5MHz.

A frequency generator will produce the three frequencies and with some other electronics we'll properly combine and feed them to the EOM.

The frequency generator will have two crystals to produce the f1 and fmc signals. The f2 modulation will be obtained by a frequency multiplier (5x) from the f1.

The frequency multiplier, for the way it works, will inevitably introduce some unwanted harmonics into the signals. These will show up as extra modulation frequencies in the EOM.

In order to quantify the effects of such unwanted harmonics on the interferometer and thus to let us set some limits on their amplitude, I ran some simulations with Optickle. The way the EOM is represented is by three RF modulators in series. In order to introduce the unwanted harmonics, I just added an RF modulator in series for each of them. I also made sure not to leave any space in between the modulators, so not to introduce phase shifts.

To check the effect at DC I looked at the sensing matrix and at the error signals. I considered the 3f error signals that we plan to use for the short DOFs and looked at how they depend on the CARM offset. I repeated the simulations for several possible amplitude of the unwanted harmonics. Some results are shown in the plots attached to this entry. 'ga' is the amplitude ratio of the unwanted
harmonics relative to the amplitude of the 11 & 55 MHz modulations.

Comparing to the case where there are no unwanted harmonics (ga = 0), one can see that not considerable effect on the error signals for amplitudes 40dB smaller than that of the main sidebands. Above that value, the REFL31I signals, that we're going to use to control PRCL, will start to be distorted: gain and linearity range change.

So 40 dB of attenuation in the unwanted harmonics is probably the minimum requirement on the frequency multiplier, although 60dB would provide a safer margin.

I'm still thinking how to evaluate any AC effect on the IFO.

 

** TODO: Plot DC sweeps with a wider range (+/- 20 pm). Also plot swept sines to look for changes in TFs out to ~10 kHz.

Attachment 1: SummaryOfResult.pdf
SummaryOfResult.pdf SummaryOfResult.pdf SummaryOfResult.pdf SummaryOfResult.pdf SummaryOfResult.pdf SummaryOfResult.pdf SummaryOfResult.pdf SummaryOfResult.pdf
  1682   Wed Jun 17 01:07:50 2009 robConfigurationComputersmatapps on /cvs/cds

I checked out a copy of matapps into /cvs/cds/caltech/apps/lscsoft so that I could find the matlab function strassign.m, which is necessary for some old mDV commands to run.  I don't know why it became necessary or why it disappeared if it did.

  1683   Wed Jun 17 01:09:47 2009 robUpdateComputers/cvs/cds 91% full

In /cvs/cds/caltech

 


1.6M    2008-8-15.pdf
2.9M    40mUpgradeOpticalLayoutPlan01.pdf
2.4M    alh
19M     apache
18G     apps
11M     archive
4.0K    authorized_keys2
8.0K    backup.notes
8.0K    backup.notes~
1.9G    build
62G     burt
47M     cds
13M     cds40m
37M     chans
70G     conlog
52K     crontab
12K     cshrc.40m
12K     cshrc.40m~
36M     diag
1.4G    dmt
8.2M    framecpp-0.2.0
1.7M    free_080730.pdf
57M     gds
9.8G    home
60K     hooks
8.0K    hosts.40m
4.0K    id_rsa
10M     iscmodeling
110M    ldg-4.7
648M    libs
4.0K    log2.txt
224K    logs
0       log.txt
238M    medm
344M    NB
148M    NB_080304
211M    NB_080307
401M    NB40
1.2G    noisebudget.071109
837M    noisebudget.bak.20060623
3.5M    oldtarget
123M    root
5.7M    savesets
208K    schematics
655M    scripts
13G     scripts_archive
1.1M    state
3.7G    svn
4.0K    svn-commit.tmp
7.3G    target
295M    target_archive
6.7M    test
72K     test.png
4.0K    tmp
8.0K    typescript
35G     users
205M    wind

  1684   Thu Jun 18 23:08:46 2009 robUpdateLockingspectrum

Here's a noise spectrum of the RSE interferometer, in anti-spring mode, with RF readout.  I'd say the calibration is "loose."

I used the Buonanno & Chen modification of the KLMTV IFO transfer functions to model the DARM opto-mechanical response.  I just guessed at the quadrature, and normalized the optical gain at the frequency of the calibration line used (927Hz, not visible on the plot).

Attachment 1: DARMnoise_929352240.png
DARMnoise_929352240.png
Attachment 2: DARMnoise_929352240.pdf
DARMnoise_929352240.pdf
  1685   Thu Jun 18 23:20:40 2009 robSummaryLSCI-Q ratio with RF detected DARM

This is a ratio of PD1_I to PD1_Q (so a ratio of the two quadratures of AS166), measured in an anti-spring state.  It's not flat because our set up has single sideband RF heterodyne detection, and using a single RF sideband as a local oscillator allows one to detect different quadratures by using different RF demodulation phases.   So the variation in frequency is actually a measure of how the frequency response of DARM changes when you vary the detection quadrature.  This measure is imperfect because it doesn't account for the effect of the DARM loop.

Even though you can choose your detection quadrature with this setup, you can't get squeezed quantum noise with a single RF sideband.  The quantum noise around the other (zero-amplitude) RF sideband still gets mixed in, and negates any squeezing benefits.

Attachment 1: IQratio.jpg
IQratio.jpg
  1686   Fri Jun 19 13:38:42 2009 AlbertoConfigurationComputerselog rebooted

Today I found the elog down, so I rebooted it following the instructions in the wiki.

  1687   Fri Jun 19 13:39:29 2009 AidanUpdateGeneralUpper limit measurement of the scatter from the eLIGO beam dumps.

 

 

I measured the scatter from the eLIGO beam dumps as best I could. The experiment setup is shown in the attached diagram. 

 

After familiarizing myself with the equipment in the morning I noticed three issues with the setup

 

1 - around the minimum scatter the back scatter from the beam dump is very susceptible to the incident angle (makes sense since the Si plate inside the beam dump at Brewster's angle when there is minimum scatter).

2 - The mirrored plug (Part 20 in D0900095) which is suppose to be used for alignment is not very effective. It moves around too much in its hole in the front face of the beam dump. Just by touching it I could make the reflected beam jump around by about 0.1 radians.

- I think to align these properly we'll have to partly assemble the dumps. If we leave off the front plate of the horn then we can measure the reflection off the Si. If we measure this with a power meter then alignment becomes a simple matter of rotating until this reflection is minimized.

3. - For this measurement the incident beam was a small (~ 1mm diameter) central beam with a small amount of spray of laser light beyond that central region. This spray was hitting the aluminium front face of the beam dump and was scattering back to the photodiode. This was clearly the limiting factor in the measurement. Most of this light was spread horizontally so I placed a couple of pieces of black glass on either side of the aperture, just blocking the edges a little. This reduce the background reading at the minimum scatter from 17.0uV to around 4.5uV with still a little bit of light hitting the top and bottom of beam dump face.

 

The incident power on the beam dump fluctuated a little but was in the range 20.5 to 22mW. The response of the PD is approximately 0.2 A/W and the transimpedance is 7.5E4 V/A.

 

The SR830 Sensitivity was set to 1x1 mV.

 

It was difficult to measure the actual angle of incidence. The dump pivoted about a point directly under the input aperture at the front. By measuring the displacement of a point on the back of the dump as I rotated it and knowing the distance between this point and the pivot point I was able to make a reasonably accurate measurement of a range of angles about the minimum.

 

The measured scatter (in V measured directly by the PD and as a fraction of the incident power) is shown in the attached plots.

 

I think I can do a better job cleaning up the incident beam - so these numbers only represent an upper limit on the scatter.

 

attachment 1: beam dump assembly

attachment 2: experimental layout

attachment 3: scatter measurement

attachment 4: BRDF - (scatter divided by the solid angle = 1.1 m steradians)

attachment 5: (slightly blurred )photo of dump - overhead view 

Attachment 1: D0900095_ELIGO_35_WATT_AIR_COOLED_BEAM_DUMP_3INCH_P.POL-3.PDF
D0900095_ELIGO_35_WATT_AIR_COOLED_BEAM_DUMP_3INCH_P.POL-3.PDF
Attachment 2: beam_dump_expt.png
beam_dump_expt.png
Attachment 3: scatter_measurement1.pdf
scatter_measurement1.pdf
Attachment 4: BRDF1.pdf
BRDF1.pdf
Attachment 5: IMG_0308.JPG
IMG_0308.JPG
  1688   Fri Jun 19 14:30:47 2009 AlbertoConfigurationComputerselog rebooted

Quote:

Today I found the elog down, so I rebooted it following the instructions in the wiki.

 I have the impression that Nodus has been rebooted since last night, hasn't it?

  1689   Sun Jun 21 00:08:26 2009 ranaConfigurationComputerselog rebooted

 

nodus:log>dmesg

Sun Jun 21 00:06:43 PDT 2009
Mar  6 15:46:32 nodus sshd[26490]: [ID 800047 auth.crit] fatal: Timeout before authentication for 131.215.114.93
Mar 10 11:11:32 nodus sshd[22775]: [ID 800047 auth.crit] fatal: Timeout before authentication for 131.215.114.93
Mar 11 13:27:37 nodus sshd[7768]: [ID 800047 auth.error] error: connect_to 131.215.115.52 port 7000: Connection refused
Mar 11 13:27:37 nodus sshd[7768]: [ID 800047 auth.error] error: connect_to nodus port 7000: failed.
Mar 11 13:31:40 nodus sshd[7768]: [ID 800047 auth.error] error: connect_to 131.215.115.52 port 7000: Connection refused
Mar 11 13:31:40 nodus sshd[7768]: [ID 800047 auth.error] error: connect_to nodus port 7000: failed.
Mar 11 13:31:45 nodus sshd[7768]: [ID 800047 auth.error] error: connect_to 131.215.115.52 port 7000: Connection refused
Mar 11 13:31:45 nodus sshd[7768]: [ID 800047 auth.error] error: connect_to nodus port 7000: failed.
Mar 11 13:34:58 nodus sshd[7768]: [ID 800047 auth.error] error: connect_to 131.215.115.52 port 7000: Connection refused
Mar 11 13:34:58 nodus sshd[7768]: [ID 800047 auth.error] error: connect_to nodus port 7000: failed.
Mar 12 16:09:23 nodus sshd[22785]: [ID 800047 auth.crit] fatal: Timeout before authentication for 131.215.114.93
Mar 14 20:14:42 nodus sshd[13563]: [ID 800047 auth.crit] fatal: Timeout before authentication for 131.215.114.93
Mar 25 19:47:19 nodus sudo: [ID 702911 local2.alert] controls : 3 incorrect password attempts ; TTY=pts/2 ; PWD=/cvs/cds ; USER=root ; COMMAND=/usr/bin/rm -rf kamioka/
Mar 25 19:48:46 nodus su: [ID 810491 auth.crit] 'su root' failed for controls on /dev/pts/2
Mar 25 19:49:17 nodus last message repeated 2 times
Mar 25 19:51:14 nodus sudo: [ID 702911 local2.alert] controls : 1 incorrect password attempt ; TTY=pts/2 ; PWD=/cvs/cds ; USER=root ; COMMAND=/usr/bin/rm -rf kamioka/
Mar 25 19:51:22 nodus su: [ID 810491 auth.crit] 'su root' failed for controls on /dev/pts/2
Jun  8 16:12:17 nodus su: [ID 810491 auth.crit] 'su root' failed for controls on /dev/pts/4

nodus:log>uptime
 12:06am  up 150 day(s), 11:52,  1 user,  load average: 0.05, 0.07, 0.07

  1690   Tue Jun 23 10:06:03 2009 steveUpdateVACnew maglev RGA scan at day6

The Maglev is running for 10 days with V1 closed.  The pressure at the RGA-region is  at 2e-9 torr on CC4 cold cathode gauge.

Valve VM2 to Rga-only was opened 6 days ago. The foreline pressure is still 2.2e-6 torr with small Varian turbo ~10 l/s on cc2

Daily scans show small improvement in large amu 32 Oxygen and large amu 16, 17 and 18 H20 water peaks.

Argon calibration valve is leaking on our Ar cylinder and it is constant.

 

The good news is that there are no fragmented hydrocarbons in the spectrum.

The Maglev is soaked with water. It was seating in the 40m for 4 years with viton o-ring seals

However I can not explan the large oxygen peak, either Rai Weiss can not.

 

The Maglev scans are indicating cleanliness and water. I'm ready to open V1 to the IFO

Attachment 1: newmagscand6.jpg
newmagscand6.jpg
Attachment 2: nmagpresd10.jpg
nmagpresd10.jpg
  1691   Tue Jun 23 15:19:42 2009 steveConfigurationVACV1 is open

Quote:

The Maglev is running for 10 days with V1 closed.  The pressure at the RGA-region is  at 2e-9 torr on CC4 cold cathode gauge.

Valve VM2 to Rga-only was opened 6 days ago. The foreline pressure is still 2.2e-6 torr with small Varian turbo ~10 l/s on cc2

Daily scans show small improvement in large amu 32 Oxygen and large amu 16, 17 and 18 H20 water peaks.

Argon calibration valve is leaking on our Ar cylinder and it is constant.

 

The good news is that there are no fragmented hydrocarbons in the spectrum.

The Maglev is soaked with water. It was seating in the 40m for 4 years with viton o-ring seals

However I can not explan the large oxygen peak, either Rai Weiss can not.

 

The Maglev scans are indicating cleanliness and water. I'm ready to open V1 to the IFO

 V1 valve is open to IFO now. V1 interlock will be tested tomorrow.

Valve configuration: VAC NORMAL with CRYO and Maglev are both pumping on the IFO

Attachment 1: V1_opennmag.jpg
V1_opennmag.jpg
  1692   Tue Jun 23 23:14:36 2009 ClaraConfigurationPEMAccelerometers relocated

Both accelerometers have been moved in an attempt to optimize their positions. The MC1 accelerometer was moved from one green bar to the other (I don't know what to call them) at the base of the MC1 and MC3 chambers. That area is pretty tight, as there is an optical table right there, and I did my best to be careful, but if you suspect something has been knocked loose, you might check in that area. The MC2 accelerometer was moved from the horizontal bar down to the metal table on which the MC2 chamber rests.

  1693   Wed Jun 24 09:21:19 2009 steveConfigurationVACIFO RGA scan with Maglev & Cryo

Quote:

Quote:

The Maglev is running for 10 days with V1 closed.  The pressure at the RGA-region is  at 2e-9 torr on CC4 cold cathode gauge.

Valve VM2 to Rga-only was opened 6 days ago. The foreline pressure is still 2.2e-6 torr with small Varian turbo ~10 l/s on cc2

Daily scans show small improvement in large amu 32 Oxygen and large amu 16, 17 and 18 H20 water peaks.

Argon calibration valve is leaking on our Ar cylinder and it is constant.

 

The good news is that there are no fragmented hydrocarbons in the spectrum.

The Maglev is soaked with water. It was seating in the 40m for 4 years with viton o-ring seals

However I can not explan the large oxygen peak, either Rai Weiss can not.

 

The Maglev scans are indicating cleanliness and water. I'm ready to open V1 to the IFO

 V1 valve is open to IFO now. V1 interlock will be tested tomorrow.

Valve configuration: VAC NORMAL with CRYO and Maglev are both pumping on the IFO

The IFO RGA scan is normal.

The Cryo needs to be regenerated next. It has been pumping for 36 days since last regenerated.

This has to be done periodically, so the Cryo's  14 K cold head is not insulated by by ice of all things pumped away from the IFO

Attachment 1: nmagd1ifo.jpg
nmagd1ifo.jpg
Attachment 2: nmagCryopres.jpg
nmagCryopres.jpg
  1694   Wed Jun 24 10:53:34 2009 Chris ZimmermanUpdateGeneralWeek 1/2 Update

I've spent most of the last week doing background reading; fourier transforms, shm, e&m, and other physics that I didn't cover at school.  I also read a few chapters in Saulson, especially the chapter on noise and shot noise.  To get a better grip on what I'm going to be doing I read through the polarization chapter in Hobbs' "Optics" text, mostly on wave plates since that's a large part of this readout.  Since then I've been working up to calculating the shot noise, starting with the electric field throughout the new interferometer readout.

  1695   Wed Jun 24 11:20:40 2009 StephanieUpdateGeneralMultiply Resonant EOM Update

I have created the attached EOM circuit with resonances at 11 MHz, 29.5 MHz, and 55 MHz (the magnitude and phase of the voltage across the EOM are shown in the attached plot). The gain is roughly the same for each resonant peak. Although I have managed to get the impedances at all of the resonant frequencies to equal each other, I am having more trouble getting the impedances to be 50 Ohms (they are currently all around 0.66 Ohms).

For the current circuit, initial calculations show that we will need around 4.7 - 14.2 A of current to drive the EOM at the desired voltage (8 - 24 V); this is much higher than the current rating of most of the available transformers (250 mA), but the necessary current will change as the impedance of the circuit is corrected, so this is probably not a cause for concern. For example, the necessary driving voltages for the current circuit are (2.8 - 8.5 V); if we assume that the 50-Ohm impedance will be purely resistive, then we get a current range of 56 - 170 mA.

Attachment 1: EOM_CktDiagram.JPG
EOM_CktDiagram.JPG
Attachment 2: EOM_VoltagePlot.JPG
EOM_VoltagePlot.JPG
  1696   Wed Jun 24 12:04:00 2009 ClaraUpdatePEMaccelerometer clarification

When I said "MC1/MC2 accelerometers," I meant the entire three-axis accelerometer set at each point.

  1697   Wed Jun 24 12:04:22 2009 ZachUpdateCamerasSURF entry

This week, I've been reading some literature concerning PLL and familiarizing myself with LINUX, MATLAB, and high-pass filter circuits.  In MATLAB, I started constructing matrices to be used for a beam path analysis from the laser output to the ccd camera.  I also built a simple high-pass filter on a bread-board that Joe and I are currently testing with the spectrum analyzer.

  1698   Wed Jun 24 12:09:24 2009 ClaraUpdatePEMWeek 1(ish)

I spent the week reading up on filter algorithm theory, particularly Wiener filtering. I have also learned how to get data from specific channels at specific times, and I've been getting myself acquainted with Matlab (which I have not previously used). Finally, I started messing around with the positioning of the accelerometers and seismometers in order to try to find the setup that yields the best filtration.

  1699   Wed Jun 24 14:43:19 2009 robUpdateComputerstdsresp on linux => pzresp

 

tdsresp is broken on our linux control room machines.  I made a little perl replacement which uses the DiagTools.pm perl module, called pzresp.  It's in the $SCRIPTS/general directory, and so in the path of all the machines.  I also edited the cshrc.40m file so that on linux machines tdsresp points to this perl replacement.

I've patched DiagTools.pm to circumnavigate the tdsdmd bug described here.  I also added a function to DiagTools.pm called diagRespNoLog, which is just like diagResp but without that pesky log file.

 


Here's the output from the tdsresp binary on CentOS:
allegra:~>tdsresp 941.54 10000 100 10 C1:LSC-ITMX_EXC C1:LSC-PD1_Q C1:LSC-PD1_I
nan nan nan nan nan nan nan
nan nan nan nan nan nan nan
nan nan nan nan nan nan nan
*** glibc detected *** tdsresp: free(): invalid next size (fast): 0x089483e8 ***
======= Backtrace: =========
/lib/libc.so.6[0xa800f1]
/lib/libc.so.6(cfree+0x90)[0xa83bc0]
/usr/lib/libstdc++.so.6(_ZdlPv+0x21)[0xf7f36571]
tdsresp[0x8057fbb]
tdsresp[0x805b394]
/lib/libc.so.6(__libc_start_main+0xdc)[0xa2ce8c]
tdsresp(__gxx_personality_v0+0x169)[0x804ddd1]
======= Memory map: ========
00242000-00249000 r-xp 00000000 fd:00 15400987                           /lib/librt-2.5.so
00249000-0024a000 r--p 00006000 fd:00 15400987                           /lib/librt-2.5.so
0024a000-0024b000 rw-p 00007000 fd:00 15400987                           /lib/librt-2.5.so
009f9000-00a13000 r-xp 00000000 fd:00 15400963                           /lib/ld-2.5.so
00a13000-00a14000 r--p 00019000 fd:00 15400963                           /lib/ld-2.5.so
00a14000-00a15000 rw-p 0001a000 fd:00 15400963                           /lib/ld-2.5.so
00a17000-00b55000 r-xp 00000000 fd:00 15400974                           /lib/libc-2.5.so
00b55000-00b57000 r--p 0013e000 fd:00 15400974                           /lib/libc-2.5.so
00b57000-00b58000 rw-p 00140000 fd:00 15400974                           /lib/libc-2.5.so
00b58000-00b5b000 rw-p 00b58000 00:00 0 
00b5d000-00b70000 r-xp 00000000 fd:00 15400984                           /lib/libpthread-2.5.so
00b70000-00b71000 r--p 00012000 fd:00 15400984                           /lib/libpthread-2.5.so
00b71000-00b72000 rw-p 00013000 fd:00 15400984                           /lib/libpthread-2.5.so
00b72000-00b74000 rw-p 00b72000 00:00 0 
00b76000-00b78000 r-xp 00000000 fd:00 15400981                           /lib/libdl-2.5.so
00b78000-00b79000 r--p 00001000 fd:00 15400981                           /lib/libdl-2.5.so
00b79000-00b7a000 rw-p 00002000 fd:00 15400981                           /lib/libdl-2.5.so
00b7c000-00ba1000 r-xp 00000000 fd:00 15400975                           /lib/libm-2.5.so
00ba1000-00ba2000 r--p 00024000 fd:00 15400975                           /lib/libm-2.5.so
00ba2000-00ba3000 rw-p 00025000 fd:00 15400975                           /lib/libm-2.5.so
00bca000-00bdd000 r-xp 00000000 fd:00 15401011                           /lib/libnsl-2.5.so
00bdd000-00bde000 r--p 00012000 fd:00 15401011                           /lib/libnsl-2.5.so
00bde000-00bdf000 rw-p 00013000 fd:00 15401011                           /lib/libnsl-2.5.so
00bdf000-00be1000 rw-p 00bdf000 00:00 0 
00dca000-00dd5000 r-xp 00000000 fd:00 15400986                           /lib/libgcc_s-4.1.2-20080825.so.1
00dd5000-00dd6000 rw-p 0000a000 fd:00 15400986                           /lib/libgcc_s-4.1.2-20080825.so.1
08048000-080b7000 r-xp 00000000 00:17 6455328                            /cvs/cds/caltech/apps/linux/tds/bin/tdsresp
080b7000-080ba000 rw-p 0006e000 00:17 6455328                            /cvs/cds/caltech/apps/linux/tds/bin/tdsresp
080ba000-080bb000 rw-p 080ba000 00:00 0 
0893d000-0896b000 rw-p 0893d000 00:00 0                                  [heap]
f5e73000-f5e74000 ---p f5e73000 00:00 0 
f5e74000-f6874000 rw-p f5e74000 00:00 0 
f692d000-f6931000 r-xp 00000000 fd:00 15400995                           /lib/libnss_dns-2.5.so
f6931000-f6932000 r--p 00003000 fd:00 15400995                           /lib/libnss_dns-2.5.so
f6932000-f6933000 rw-p 00004000 fd:00 15400995                           /lib/libnss_dns-2.5.so
f6956000-f6a12000 rw-p f6a31000 00:00 0 
f6a74000-f6a7d000 r-xp 00000000 fd:00 15400997                           /lib/libnss_files-2.5.so
f6a7d000-f6a7e000 r--p 00008000 fd:00 15400997                           /lib/libnss_files-2.5.so
f6a7e000-f6a7f000 rw-p 00009000 fd:00 15400997                           /lib/libnss_files-2.5.so
f6a7f000-f6a80000 ---p f6a7f000 00:00 0 
f6a80000-f7480000 rw-p f6a80000 00:00 0 
f7480000-f7481000 ---p f7480000 00:00 0 
f7481000-f7e83000 rw-p f7481000 00:00 0 
f7e83000-f7f63000 r-xp 00000000 fd:00 6236924                            /usr/lib/libstdc++.so.6.0.8
f7f63000-f7f67000 r--p 000df000 fd:00 6236924                            /usr/libAbort

 
  1700   Wed Jun 24 17:08:16 2009 steveBureaucracySAFETYsurf student safety

The 40m lab specific safety training is done. The participants were

Stephanie Erickson, Clara Bennett, Chris Zimmerman, Zach Commings, Michelle Stephen surfs and Drew Cappel  postock.

They have already went through the Caltech Safety Office laser and general safety training.

They still have to read, understand and sign the the SOP for the laser & lab

  1701   Thu Jun 25 10:28:58 2009 steveHowToVACCryopump is regenerated
The Cryopump's VC1 valve to IFO was closed yesterday.
The compressor  Helium pressure was 235 PSI. The cold head temp on H2 vapor pressure gauge was reading ~14 Kelvin,
The compressor and piston driver were turned off to let cold head warm up to room temp.
The flow path from Cryo to TP3 were checked to insure that only VC2 and V5 would be open for pumping.
VC2 valve was opened to TP3 through V5
Now as the Cryo was warming up while TP3 drag turbo pump was pumping away the accumulated ice, that was melting and vaporizing.
This is shown on one day the plot below.
To check outgassing rate of the Cryo pump after one day of pumping V5 was closed for 20 minutes.
The accumulation was 1.3 mTorr in 20 min
This means the Cryo is clean, it is ready to be started up in the future.
VC2 was closed to seal this condition.
The flow path between VC2, VM3, V7, V6 , VA6 and manual needle valve  would be pumped for one day through V5 to TP3 to clean up 
Attachment 1: Creg.jpg
Creg.jpg
  1702   Thu Jun 25 17:27:42 2009 ranaUpdateComputerstdsresp on linux
I downloaded the tdsresp.pl from LLO and put it into the various TDS/bin paths. Also updated the LLO specific path stuff. It runs.
  1703   Thu Jun 25 21:00:30 2009 ClaraUpdatePEMMC1 Accelerator set moved again; new XLR cables

I moved the MC1 set of accelerators. Might have bumped things. If things aren't working, look around the MC1 chamber.

Also, I constructed two new XLR cables, but have not tested them yet.

  1704   Fri Jun 26 15:22:28 2009 ClaraUpdatePEMXLR cables tested and labeled

We now have two 80-foot, female-to-female XLR cables for our pretty new microphones, one yellow and one purple. They have been tested and appropriately labeled.

Also, here is a very helpful pdf for how to properly attach the XLR connectors to a raw quad cable, as well as one for how to put the actual connectors together (ignore the cable instructions on the connector page... the cable depicted is not a quad cable).

 

 

Attachment 1: NC3FXX-EMC.pdf
NC3FXX-EMC.pdf
Attachment 2: Cat11_p35.pdf
Cat11_p35.pdf
  1705   Fri Jun 26 18:00:36 2009 JenneUpdatePEMNew PEM channels for the fancy-pants new microphones

[ Jenne, Clara ]

We made new channels for the microphones which came in this week, by editing C1ADCU_PEM.ini (and making an appropriate backup before we modified it) then restarting the framebuilder and the frontend computer C0DCU1.  The new channels are:

C1:PEM-AUDIO_MIC1

C1:PEM-AUDIO_MIC2

These are connected to channels 13 and 14 on the PEM ADCU board, just next to the GURALP seismometer channels.

 

Clara is testing the mics so the max output voltage can be limited to +-2V for the DAQ, then we'll hook them up to our new channels and listen to the IFO (and all the audio frequency noises around it).

  1706   Fri Jun 26 19:14:04 2009 ranaUpdateEnvironmentseisBLRMS for the past 3 weeks
Restarted the seisBLRMS.m on mafalda (running a term on op540m). Don't know why it stopped - the
terminal had a 'disabled by EPICS' message even though the EPICS enable button was enabled.

I also changed the delay from 4 to 2 minutes. So now it is calculating a 64 s PSD starting from 2 minutes ago. We had
put it back to 4 minutes long ago since the framebuilder was acting slow, but maybe it will work now since its using
the NDS1 protocol instead of direct frame reading. If it starts not finding data, please kill the seisBLRMS and restart
it with a 3 minute delay.
  1707   Sat Jun 27 02:48:09 2009 ClaraUpdatePEMI moved accelerometers and made some pretty pictures

I have been working on finding the best spots to put the accelerometer sets in order to best subtract out noise (seismometers next!). Here is a plot of what I've done so far:

80min_accel_0123.png

All of these were 80-minute samples. The dashed line is unfiltered, solid line filtered. So, Setup #1 looks the best so far, but I didn't leave it there very long, so perhaps it was just a really awesome 80 minutes. I've put the accelerometers back in the Setup #1 position to make sure that it is really better.

And, in case you can't intuitively figure out what configuration the accelerometers are in by such descriptive names, here are some helpful pictures. I didn't know about the digital cameras at first, so these are actually sketches from my notebook, which I helpfully labeled with the setup numbers, color-coded to match the graph above! Also, there are some real-life photographs of the current arrangement (Setup #1' if you forgot).

MC1_accel_sketch_side.png

MC1_accel_sketch_top.png

MC2_accel_sketch.png

Doesn't this one look kind of Quentin Blake-esque? (He illustrated for Roald Dahl.)

MC1acc_S01.JPG

This is the MC1 set.

MC2acc_S123.JPG

Guess which one this is!

  1708   Sat Jun 27 03:16:16 2009 ClaraUpdatePEMExciting microphone things!

So, I'm double-posting, but I figured the last post was long enough as it was, and this is about something different. After double and triple checking the XLR cables, I hooked up the microphone setup (mic---preamp---output) to the oscilloscope to figure out what kind of voltage would register with loud noises. So, I clapped and shouted and forgot to warn the other people in the lab what I was doing (sorry guys) and discovered that, even on the lowest gain setting, my loud noises were generation 2-3 times as much voltage as the ADC can handle (2V). And, since our XLR cables are so freaking long, we probably want to go for a higher gain, which puts us at something like 20 times too much voltage. I doubt this is really necessary, but it's late (early) and I got camera-happy, so I'm going to share anyway:

mic_voltage_out.png

So, to deal with this issue, I made some nifty voltage dividers. Hopefully they are small enough to fit side-by-side in the ports without needing extra cableage. Anyway, they should prevent the voltage from getting larger than 2V at the output even if the mic setup is producing 50V. Seeing as my screaming as loud as I could about 2mm away from the mic at full gain could only produce 45V, I think this should be pretty safe. I put the ADC in parallel with a 25.5 kOhm resistor, which should have a noise like 10^-8 V/rHz. This is a lot smaller than 1 uV/rHz (the noise in the ADC, if I understood Rana's explanation correctly), so the voltage dividers should pose a noise issue. Now for pictures.

voltage_dividers.png

I opened one so you can see its innards.

voltage_divider_sketch.png

In case the diagram on the box was too small to decipher...

And finally, I came up with a name scheme for the mics and pre-amps. We now have two Bluebird (bacteriophage) mics named Bonnie and Butch Cassidy. Their preamps are, naturally, Clyde and The Sundance Kid. Sadly, no photos. I know it's disappointing. Also, before anyone gives me crap for putting the labels on the mics upside-down, they are meant to be hung or mounted from high things, and the location (and stiffness) of the cable prevents us from simply standing them up. So they will more than likely be in some kind of upsidedownish position.

  1709   Tue Jun 30 23:09:40 2009 AlbertoUpdateLockingchronicles of some locking attempts

Tonight I tried to lock the interferometer. At the first attempts the arm power didn't go above about 4. The mode cleaner seemed to be not well aligned and it lost lock or got stuck on a wrong mode. I had to run the MC_UP and MC_DOWN scripts to lock it again.

After that the locking proceed more smoothly; at least till a power level in the arms of about 60. Then again the mode cleaner lost lock and I had to run the scripts again. Without the MCWFS servo off the MC reflected power is still rather high (about 1.7); also even when the WFS servo is engaged the reflected power is about 0.5, versus 0.3 that it should be.

Those are both signs of a not very good alignment. Tomorrow I'll have to work on the injection periscope on the PSL table to try to fix that.

  1710   Wed Jul 1 10:56:42 2009 Chris ZimmermanUpdateGeneralWeek 2/3 Update

I spent the last week working a lot with the differences between a basic Michelson readout and the new one as a displacement sensor.  The new one (w/ wave plates) ends with two differently polarized beams and should have better sensitivity; I've also been going through noise/sensitivity calculations for each, although that hit a road block when I had to start the 1st SURF progress report, which has taken up most of my time since Saturday.

  1711   Wed Jul 1 11:00:52 2009 StephanieUpdateGeneralMultiply Resonant EOM Update

Since last week, I have come up with a new circuit, which is shown in the attached figure. The magnitude (solid) and phase (dashed) of the voltage across the EOM (red), the ratio between the voltage across the EOM and the voltage across the primary nodes of the transformer (blue), and the impedance through the primary port of the transformer (red) are also shown in an attached figure. As can be seen on the plot, resonance occurs at 11 MHz, 29.5 MHz, and 55 MHz, as specified. Also, at these resonant frequencies, the impedance is about 50 Ohms (34 dB). The gain between the voltage across the EOM and the voltage across the primary nodes of the transformer (or output of the crystal oscillator) is about 12 dB; we'd like a higher gain than this, but this gain is primarily governed by the ratio between the secondary and primary inductances in the transformer, and we are using the largest available ratio (on the Coilcraft website) that has the necessary bandwidth. Because of this, we will likely have to add another component between the crystal oscillator and the EOM circuit, to get the voltage to the desired 8.5 Vp across the EOM (for an optical modulation depth of 0.1 rad).

The current and power through the primary port of the tranformer are 43-85 mA and 25-92 mW, respectively. Since the transformer ratings are 250 mA and 1/4 W for current and power; these values should be safe to use with the intended transformer. Also, the highest power dissipated by a resistor in the circuit (not including the 50 Ohm resistor that is part of the crystal oscillator setup) is around 74 mW.

Attachment 1: EOMCKT.png
EOMCKT.png
Attachment 2: PR1_VoltagePlot.pdf
PR1_VoltagePlot.pdf
  1712   Wed Jul 1 11:04:27 2009 ZachUpdateCamerasGigE Phase Camera

This past week, I have building a sine wave rectifier and trying to write a simple program that displays a ccd image to familiarize myself with the code.  I also wrote a progress report in which I included the following images of the sine wave rectifier circuit as well as the optical chain including the phase-locked loop.  The hirose connector arrived so I can begin soldering the electronics together and testing the trigger box with the ccd.  I am waiting on the universal PDH box as well as another fiber coupler to begin setting up the optics.  In order to avoid the frustrations associated with sending a laser beam down a long pipe to an optical bench across the room, I will be transmitting laser 1 to the ccd by means of a fiber optic cable and dealing with the alternative new and exciting frustrations.

 

Attachment 1: trigger.jpg
trigger.jpg
Attachment 2: fig1b.pdf
fig1b.pdf
  1713   Thu Jul 2 05:27:12 2009 ClaraUpdatePEMVoltage Divider Oops

I tested the voltage dividers and was getting up to about 3V. I retested the mic w/o the voltage divider in place, and, lo and behold, I was able to generate about 70-75V (previously, I maxed out at 45V). I'm not 100% sure why this was, but it occurs to me that, before, the sounds I was generating were short in duration (loud claps, short yelps). This time, I tried yelling continuously into the microphone. So, probably, I simply wasn't seeing the real peak before on the scope because it was too short to pick up. I have corrected the voltage dividers (by replacing the 25.5 kOhm resistors, which were in parallel with the ADC, with 10 kOhm resistors, taking the voltage ratio to ~60:1) and tested them. I haven't been able to generate more than 1500 mV, so I think they are safe. (It's possible we would have been fine with the old setup, since I think it would be hard to get any noises as loud as I was making, but better safe than sorry, right?)

I'm attaching a diagram of the new-and-improved voltage dividers.

voltage_divider_diagram.png

  1714   Thu Jul 2 06:31:35 2009 ClaraUpdatePEMFirst mic in place and connected

I clamped Bonnie (microphone) to the top of a chamber near the vertex of the arms and placed Clyde (pre-amp) on the table right below (see picture). The cable was laid and Bonnie and Clyde are plugged into port #13 on the ADC. The second cable was plugged into port #14, but it is not connected to anything. I placed the looped up cable on top of the cabinet holding the ADC.

Note: the angle in the photograph is such that we are looking along the y-arm.


Attachment 1: bonnieandclyde.JPG
bonnieandclyde.JPG