  170   Wed Dec 5 19:25:07 2007 ranaDAQCDSDMF
I made a database file on C1AUX called dmf.db. It has 9 DMF EPICS channels which are also trended
so that one can now write data to those channels from a DMF Monitor and the data will be records.

New channels:
[C1:DMF-SEIS_1]
[C1:DMF-SEIS_2]
[C1:DMF-SEIS_3]
[C1:DMF-LINE_1]
[C1:DMF-LINE_2]
[C1:DMF-LINE_3]
[C1:DMF-MC_1]
[C1:DMF-MC_2]
[C1:DMF-MC_3]

I added these to C1AUX because it doesn't do much and can be booted without having much effect.
(it controls Mech Shutters, Video, and Illuminators. It used to also do the EO Shutter but I
removed that from its startup.cmd and it will no longer load those records).
  169   Wed Dec 5 18:22:03 2007 tobinUpdatePSLISS dark noise
Attached is a plot of the dark noise spectrum of the ISS photodiodes (1) before fooling with them, (2) after replacing the 4151's with 4131's (improvement!), and (3) after replacing the cables and changing the wiring (disaster!).
  168   Wed Dec 5 18:08:36 2007 AndreyUpdateASCOptical Lever laser for ETMX is installed

A new laser with \lambda=633nm has been intalled and the mirror adjusted so that the signal hits the center of the photodetector.

Output power level of that laser is 3.45 +- 0.05 mW.

Only about 0.29mW hits the photodetector.

Cable clips have been used to firmly fix the power supply cable for the laser.

See attached photopicture of the ETMX - "oplev" - optical - table.
  167   Wed Dec 5 17:49:57 2007 tobinUpdatePSLISS
Attached is a plot of the ISS RIN with a variety of gain settings.

Unfortunately the dark noise is huge now--a result of the new cables & wiring?
  166   Wed Dec 5 16:57:36 2007 tobinHowToComputer Scripts / ProgramsSR785 data converter on linux
I was pleased to find that the SR785 Data Viewer (including the command line conversion utility) installs and works in linux using WINE (on my laptop at least). There are some quirks, of course, but I was able to extract my data.
  165   Wed Dec 5 13:49:08 2007 albertoUpdateElectronicsRF AM PD lines monitor
In the last weeks Iíve been working on the design of an electronic board to measure directly the power of the main spectral lines on of the RF-AM photodiode from as many independent outputs. The idea is to have eventually a monitor channel in the CDS network for the power of each line.
Looking at at the spectrum from the RF-AM PD (see attached plot), there are 5 main lines:
Frequency
3 fsr = 33 195 439 Hz
4 fsr = 66 390 878 Hz
12 fsr = 132 781 756 Hz
15 fsr = 165 977 195 Hz
18 fsr = 199 172 634 Hz

Two main approaches have been proposed for the circuit depending on the way followed to isolate the lines:
1) Filters: the frequencies are separated by narrow notch filters, then a diode bridge rectifies and a low pass filter extracts the DC component.
2) Mixers: for each frequency there is a mixer driven by a copy of the correspondent modulation frequency provided by the function generators (the Marconi). The mixers automatically give the DC component of the rectified signals.
Because of the phase lags that we should compensate if we used mixers, we would prefer the first approach, if it works.
Starting with a tolerance of about 10% between the channels, the spectrum (see attachment) sets the constraint to the filterís suppression:
Filter central frequency [MHz]******Suppression within 30 Mhz [dB]
33*********************************-7-20 = -27
66**********************************7-20 = -13
133*********************************12-20 = -8
166********************************-12-20 = -32
199*********************************10-20 = 10

So far Iíve tried two kinds of designs for the filters, Butterworth (see attachment) and LC and I'm measuring transfer functions tuning the components to match the central frequency and the bandwdth of the filters with the requirements.

The frequencies weíre dealing with are rather high and several adjustments had to be done to the measurement system in order to shield the circuit from the impedance of the input and the output line (i.e., amplifier turned out to be necessary). Also, an the mixer had to be replaced to an RF one.
It seems I'm now measuring new transfer functions (which look quite different from what I've got with no amplifiers).
To be posted soon.
  164   Wed Dec 5 10:57:08 2007 albertoHowToComputersConnecting the GPIBto USB interface to the Dell laptop
The interface works only on one of the USB ports of the laptop (the one on the right, looking at the computer from the back).
  163   Tue Dec 4 23:16:35 2007 tobinUpdatePSLISS
I was confused to find that I could increase the ISS gain slider all the way from 15dB to 30dB without seeing much of any increase in gain in the measured open-loop transfer function. While making these swept-sine measurements, the saturation indicator almost never tripped, indicating it was seemingly happy. But then I noticed an odd thing: if I disable the test ("analog excitation") input, the saturation indicator trips immediately. I hooked up a scope to the current shunt test point (TP12). With the test input enabled, the loop closed, and the analog excitation port connected to the SR785, I see a a 5 Vpkpk, 2.55 MHz triangle wave there. It is there even if I set the SR785 excitation amplitude to zero, but it disappears if I disconnect the cable from the SR785.

I found oscillations at TP20, TP30, TP36, TP41, and TP42. Many of these are in the (unused) "outer loop" circuitry and currently lack compensation capacitors.
  162   Mon Dec 3 22:20:09 2007 tobinConfigurationPSLISS
I replaced the painfully short 1' cables on the ISS photodiodes with luxurious five foot cables, made by chopping a ten foot Amphenol cable (P/N:CS-DSPMDB09MM-010) in half and using each half for one of the diodes. All of the ISS GND connections are wired to the PD GND, as is the PD- differential signal. The diodes are installed on the PSL table, but I have not tested them beyond looking at the DC values as I blocked/unblocked the beam.
  161   Mon Dec 3 19:44:58 2007 Accelerometers on new mountsConfigurationPEMAndrey

I (Andrey) continued today working with new accelerometer mounting. (see entry #151 about my Friday work).

I bought screws/washers and attached those mounts with accelerometers to metallic frames which are firmly cemented to the floor.

One such mount with three accelerometers (in X-, Y-, Z-directions) is installed near the ITMX (in the previous location, but NOT on top of the unused stack as before Friday), the other mount with three accelerometers in three orthogonal directions is installed near ETMX in the east end of the room (this set of accelerometers was installed between MC and BS before Friday). I uncoiled the cables, put them into the cable tray towards the ETMX, and hooked-up the three accelerometers near ETMX in the east end of the room.

Now all six accelerometers are hooked-up (that is, connected to power supply board with cables).

We decided with Steve Vass to put red cones (similar to those that are on highways in the road construction zones) in order to prevent people from bumping into accelerometers. Please use caution when walking along the X-arm.

I took several pictures of the new accelerometer setup. Picture "DSC_0194.JPG" shows the mount with accelerometers near the the ITMX and the beamsplitter chamber,
picture "DSC_0195.JPG" is the "zoomed-in" view of the same accelerometers, while picture "DSC_0196.JPG" shows the mount with accelerometers near ETMX in the east end of the room.

Many thanks to Mr. Steve Vass for his thorough explanation/showing me how to drill the metal and put threads in the holes.
  160   Mon Dec 3 19:06:49 2007 ranaDAQComputer Scripts / Programslinemon
I turned up my nose at Matlab's special tools. I modified the linetracker to use the
relationship phase = 2*pi*f*t to estimate the frequency each minute. The
code uses 'polyfit' to get the mean and trend of the unwrapped phase and then determines
how far the initial frequency estimate was off. It then uses the updated number as the
initial guess for the next minute.

I looked at a couple hours of data before letting it run. It looks like the phase of the
'60 Hz' peak varies at 20 second time scales but not much faster or rather anything faster
would be a glitch and not a monotonic frequency drift.

From the attached snapshot you can see that the amplitude (PZT1_PIT) varies by ~10 %
and the frequency by ~40 mHz in a couple hour span.
  159   Mon Dec 3 17:55:39 2007 tobinHowToComputer Scripts / Programslinemon
Matlab's Signal Processing toolbox has a set of algorithms for identifying sinusoids in data. Some of them (e.g., rootmusic) take the number of sinusoids to find as an argument and return the "most probable N frequencies." These could be useful in line monitoring.
  158   Mon Dec 3 16:24:47 2007 tobinHowToComputersGNU screen
GNU screen is a utility that can be quite handy for managing long-running psuedo-interactive terminal programs on remote machines. In particular, I think it might be useful in developing and testing "Matlab DMT" tools on Mafalda.
  157   Mon Dec 3 00:10:42 2007 ranaDAQComputer Scripts / Programslinemon
I've started up one of our first Matlab based DMT processes as a test.

There's a matlab script running on Mafalda which is measuring the height of the 60 Hz peak
in the MC1 UL SENSOR and writing it to an unused EPICS channel (PZT1_PIT_OFFSET).

The purpose of this is just to see if such a thing is stable over long periods of time. Its
open on a terminal on linux3 so it can be killed at any time if it runs amok.

Right now the code just demods the channel and tracks the absolute value of the peak. The
next upgrade will have it track the actual frequency once per minute and then report that
as well. We also have to figure out how to make it a binary and then make a single script
that launches all of the binaries.

For now you can watch its progress on the StripTool on op540m; its cheap and easy DMT viewer.
  156   Sun Dec 2 21:13:16 2007 ranaConfigurationIOOMC SUS re-alignment
  155   Sun Dec 2 21:07:39 2007 ranaConfigurationIOOMC SUS re-alignment
you asked for:   diff 2007/12/01,4:58:48 2007/12/03,4:58:48 utc 'MC.*COMM'
LIGO controls: differences, 2007 12/01 04:58:48 utc vs. 2007 12/03 04:58:48 utc
__Epics_Channel_Name______   __Description__________   __value1____     __value2____
C1:SUS-MC1_YAW_COMM                                    -0.273460        -0.503460
C1:SUS-MC2_PIT_COMM                                     3.624020         3.632020
C1:SUS-MC2_YAW_COMM                                    -0.936800        -1.038800
C1:SUS-MC3_YAW_COMM                                    -3.129000        -3.369000
  154   Sun Dec 2 21:02:12 2007 ranaConfigurationIOOMC SUS re-alignment
The spot on MC2 was not centered, so I put it back in the center:

  • Made sure MC trans was high with the WFS off.
  • Moved the Sliders on the MC Align screen until spot was centered (by eye)
  • Moved some more until power was maximized.
  • Unlock MC
  • Center spots on McWFS
  • Re-enable autolocker and McWFS loops.
  153   Sun Dec 2 17:37:33 2007 ranaOmnistructureComputersNetwork Cabling in the Office
We all know that we've spent many integrated man hours trying to figure out why our network connections
in the office area don't work. Usually its because of the bad hub around the Tobin/Osamu desk.

I pried open some of the wall conduit today and it looks pretty easy to fish cables through. I think
its time we finally did that. It may be a little disruptive, but I propose we get Larry to come over
and figure out what needs to happen for us to get regular 100 Mbit ports on the walls. These can
then all go over and get connected to a switch in the rack that holds linux1.

Opinions / comments ?
  152   Fri Nov 30 21:27:24 2007 ranaDAQPEMweather / stacis / c1pem1
I was trying to add some Seis BLRMS channels to the c1pem1 processor so that we could have DMT trends.

Then I found that none of the Weather channels have been working for a year or so. I could also not
telnet into it. I tried resetting it but no luck. There was no entry in the Wiki for it so I added
a place holder.

Have the weather channels ever worked? Do we have those sensors? I think I've never actually looked
for this. Seems like a fine ugrad job.
  151   Fri Nov 30 20:17:26 2007 AndreyConfigurationPEMAccelerometers and alum.plates for them
All 6 accelerometers which were located near the ITMX are turned off and disconnected from the power cords.
Actually these accelerometers are now in the office area on the electronics bench (to the left from Steve Vass' place).

I made today 4 new aluminum mounting plates for the accelerometers (I drilled holes and made threads in them). On Monday I will buy short screws and install accelerometers on these new mounting plates. These mounting plates will be screwed directly into the metallic frame which is firmly cemented to the ground. Before yesterday accelerometers were mounted on top of blue stack towers, not on the ground directly, so we hope that new measurements of the ground noise will be more realistic.

The 4 mounting plates are on the same desk -> on the electronics bench (to the left from Steve Vass' place). Please do not displace them.

Attached is a drawing of the aluminum mountain plate.
  150   Fri Nov 30 20:13:57 2007 dmassSummaryGeneralHeNe UniPhase Laser
Data for the Uniphase 1.9 mW HeNe laser (labeled: "051507 From ISCT-BS") SN: 1284131 Model: 1103P

I used the Photon Beamscanner to obtain all data, then fit w(z) as shown on the plot with parameters w_0, z_R, and hidden parameter delta,
where z = delta + x, z is waist distance, x is distance from the laser.

Copies of the matlab code used to fit (/plot) are attached in .zip below.
  149   Fri Nov 30 19:46:58 2007 ranaConfigurationComputersEPICS Time Bad again
The time on the EPICS screens is off by 10 minutes again. Por Que?

Its because the ntpd on scipe25 wasn't restarted after the last boot. If someone
knows how to put the ntpd startup into that machine, please do so.

This time I started it up by just going sshing in as controls and then entering:

sudo /usr/sbin/ntpd -c /etc/ntp.conf

which runs it as root and points to the right file.

It takes a few minutes to get going because all of the martian machines have to first fail to
connect to the worldwide pool servers (e.g. 0.pool.ntp.org) before they move on and try linux1
which has a connection to the world. Once it gets it you'll see the time on the EPICS screens
freeze. It then waits until the ntp time catches up with its old, wrong time before updating
again.

According to the Wikipedia, this time is then good to 128 ms or less.
  148   Fri Nov 30 19:29:14 2007 ranaConfigurationSUSnew screen
Andrey is working on a new screen to show us the drift of the optics by alarming on
their osem values. You can find it under SUS as 'Drift Mon' from the site map.

To aid in this I ran the following csh commands which effect all optics:
foreach opt (ETMX ETMY ITMX ITMY MC1 MC2 MC3 BS PRM SRM)
  foreach dof (POS PIT YAW)
     ezcawrite C1:SUS-${opt}_SUS${dof}_INMON.PREC 0
  end
end

This should make the DOF readouts more readable.
  147   Fri Nov 30 19:11:05 2007 ranaConfigurationElectronicsETMX oplev dead again
I removed the ETMX HeNe and put in on a test table and it fired up fine. In its
previous location the light on the HeNe power supply was not lighting up. If
that's still on over the weekend we'l blame the power strip; the HeNe is a JDS
2.7 mW laser from 2002.
  146   Fri Nov 30 13:46:50 2007 robConfigurationElectronicsETMX oplev dead again

Quote:
In the interests of getting the Xarm alignment script working again, I reset the local damping gains for the test masses to their previous known working values (1), then I noticed that the ETMX oplev was dead. Since the scripts use the oplev motion as a readback for the optic motion, this means the script was basically blindly swinging the optics around. Some monkeying around with swapping HeNe power supplies eventually led to the conclusion that the power strip is funky, since the laser works when plugged into another power strip. Even weirder, the HeNe and the power supply indicator light have some sort of XOR relationship going on. When one works, the other doesn't. Steve will sort out this confusion later; we're good for now.


Ech. The HeNe quit again. Let's replace it and see what happens.
  145   Fri Nov 30 11:44:57 2007 robConfigurationElectronicsETMX oplev
In the interests of getting the Xarm alignment script working again, I reset the local damping gains for the test masses to their previous known working values (1), then I noticed that the ETMX oplev was dead. Since the scripts use the oplev motion as a readback for the optic motion, this means the script was basically blindly swinging the optics around. Some monkeying around with swapping HeNe power supplies eventually led to the conclusion that the power strip is funky, since the laser works when plugged into another power strip. Even weirder, the HeNe and the power supply indicator light have some sort of XOR relationship going on. When one works, the other doesn't. Steve will sort out this confusion later; we're good for now.
  144   Fri Nov 30 11:22:22 2007 ajwSummaryCDSGEO DV => LIGO DV

Quote:
Martin Hewitson of GEO600 fame has modified the cool GEO DV
to work with the LIGO NDS system with some NDS advice from Rolf (who's over in Germany this week).

I've moved it onto the 40m CDS system and installed it on the AdhikariLab computer named 'django'. It worked immediately.

I modified the main .m file to include the 40m's NDS server. When you run it you have to include the path to the NDS
client written by Ben Johnson.

The attached is a screenshot of it working on a Mac; it looks as cool on Linux.

Its installed in /cvs/cds/caltech/apps/ligoDV/. In matlab you navigate to that directory and then
type addpath('/cvs/cds/caltech/apps/linux/UNIX_NDS_Client_beta2/') to add the NDS client.
On the Solaris machines, type type addpath('/cvs/cds/caltech/apps/solaris9/UNIX_NDS_Client_beta2/') instead.

Then type ligoDV to start it up. Then click away and have fun.

In the example I've selected
C1:PEM-BS_ACC_EAST_Z
and plotted its specgram.

Big grin


Download and installation instructions, as well as a few examples for use
can be found here (typical lsc username and password):

https://www.gravity.phy.syr.edu/dokuwiki/doku.php?id=ligodv:home
https://www.gravity.phy.syr.edu/dokuwiki/doku.php?id=ligodv:downloading_the_ligodv_software
  143   Thu Nov 29 19:35:14 2007 ranaHowToComputer Scripts / ProgramsGPIB Scripts

Quote:
I've spent a lot of time trying to configurate the GPIB-USB interface for the HP4195. After installing 1) the Agilent libraries, 2) the drivers, 3) the matlab Instrument Toolbox, 4) Jamie script, 5) Alice's script the computer can see the HP but still they can't 'talk' to each other.
I give up. I asked Alice Wang how she managed to get data. I'm not sure she used the GPIB interace. Rob said she might have used the old fashion floppy disks that we can't read anymore here.
I would really appreciate any suggestion by anyone who happened to have the same problems.


Alice and Jamie used the USB-GPIB interface. You should just try using the black laptop which already has this capability or ask Jamie Rollins
who actually knows something.
  142   Thu Nov 29 18:10:13 2007 AlbertoHowToComputer Scripts / ProgramsGPIB Scripts
I've spent a lot of time trying to configurate the GPIB-USB interface for the HP4195. After installing 1) the Agilent libraries, 2) the drivers, 3) the matlab Instrument Toolbox, 4) Jamie script, 5) Alice's script the computer can see the HP but still they can't 'talk' to each other.
I give up. I asked Alice Wang how she managed to get data. I'm not sure she used the GPIB interace. Rob said she might have used the old fashion floppy disks that we can't read anymore here.
I would really appreciate any suggestion by anyone who happened to have the same problems.
  141   Thu Nov 29 15:17:53 2007 robConfigurationPSLISS

I put some ISS beam on the diode on the PSL table. In the previous layout, this was the monitor diode (and it's labeled monitor) but I plugged it into the sensor jack anyways so we can run with the loop closed for now; we can just switch the cables later. The reason the beam was unclear is because someone popped up a flipper mirror which redirects the beam from the ISS into an OSA.

With the ISS gain slider at 15 dB the UGF is around 40kHz.

Why do we have such short cables for the ISS diodes?
  140   Thu Nov 29 14:29:22 2007 tobinConfigurationComputerslinux1 httpd/conlogger fixed
I think I fixed the conlogger web interface on linux1.

Steps necessary to do this:
0. Run "/etc/init.d/httpd start" to start up httpd right now
1. Run "/usr/sbin/ntsysv" and configure httpd to be started automatically in the future
2. Copy /cvs/cds/caltech/conlogger/bin/conlog_web.pl to /var/www/cgi-bin and chown to controls
8. Hack the conlog_web.pl to (0) use /usr/bin/perl (1) not use Apache::Util, and (2) function with the newer version of CGI.pm
9. Enjoy!

The following steps are optional, and may be inserted between steps 2 and 8:
3. Try to install Apache::Util (via "perl -MCPAN -e shell" followed by "Install Apache::Util")
4. Notice that the installation dies because there is no C compiler installed
5. Bang head in disgust and abomination over a Linux distribution shipping without a C compiler installed by default
6. "yum install gcc"
7. Annoyed by further dependencies, go to step 8
  139   Thu Nov 29 11:10:54 2007 robOmnistructureVACRGAlogger sleeping
Without the RGA controller responding, the RGAlogger script just hangs. Rather than fix it, I just put it to sleep by commenting out the line in op440m crontab file. Once we get it running again, we'll move the cronjob to op340m.
  138   Thu Nov 29 10:36:47 2007 albertoConfigurationComputer Scripts / ProgramsAgilent 82357B GPIB to USB Interface Installation Procedeure
To run the Agilent Automation-Ready CD provided with the interface is only the first step of the installation. Apparently there should be also a second CD with the drivers for Windows XP but I couldn't find it. So, after Installaing the IO Libraries Suite from the CD, I had to install the drivers with an executable downloaded from the Agilent's website at:

http://www.home.agilent.com/agilent/editorial.jspx?cc=US&lc=eng&ckey=1188958&nid=-35199.0.00&id=1188958

and only then I could plug in the interface.
Anyway, I burned a cd with the file and put it together with the other one.
  137   Wed Nov 28 21:51:52 2007 tobinConfigurationPSLISS
I replaced the front-end differential receivers for the ISS's "inner-loop" sensor and monitor diode inputs with lower-noise THS4131's (formerly THS4151's). I verified operation by taking the transfer function from the "PD+" and "PD-" inputs (separately) to the testpoint following the differential receiver; the surgery appears successful.

I measured the dark spectra at the ISS's DC PD BNC ports and found a noise floor of ~ 16 nV/rtHz, compared with a floor of ~ 22 nV/rtHz last week. This seems to add up, assuming the DC PD port has 0dB gain: the 4131 has a rated noise of 1.3 nV/rtHz and the 4151 a noise floor of 7.6 nV/rtHz, a difference of 6 nV/rtHz. The other change made in that time was to add a larger power supply bypass capacitor in the PD.

There are two of the old 4151 chips still on the ISS board on the two "outer-loop" channels that we don't use. If I dig up any more 5131's I will replace these too for completeness.

There is currently no light on the ISS diodes; I'm not sure where it's intended to come from.
  136   Wed Nov 28 19:44:18 2007 tobinUpdatePSLHEPA
I found the HEPA turned off completely. I turned it on.
  135   Wed Nov 28 19:02:41 2007 AndreyBureaucracyWIKI-40M UpdateNew WIKI-40M page describing Matlab Suspension Modeling

I created the WIKI-40m page with some details about my today's talk on the 40-m lab meeting.

The address is:

http://lhocds.ligo-wa.caltech.edu:8000/40m/Modeling_of_suspensions

(or you can go to the main page, http://lhocds.ligo-wa.caltech.edu:8000/40m/ , and click on the link "Modeling of suspensions").

The WIKI-40m page describes my transfer functions and contains the pdf-file of my presentation.
  134   Wed Nov 28 17:41:34 2007 robUpdatePSLFSS again
I investigated the FSS a bit more today. I looked at the signals coming out of the FSS frequency reference, and saw that both the LO and PC drive were distorted, non-symmetric waveforms. In addition, the LO path had a 3dB attenuator, meaning the mixer was starved. I placed mini-circuits SLP-30 filters in both paths, and now both are nice sine waves. I also took out the 3dB att. With this work, and the CG slider maxed out at 30, the FSS open loop gain (for real this time) goes up to ~250kHz. Still needs more investigation.
  133   Wed Nov 28 17:15:26 2007 ranaConfigurationSUSETMY damping / watchdogs
Steve has noted that ETMY was often tripping its watchdog. I saw this again today.

So I checked the damping settings. Someone had set the SIDE gain to +1. The gain which gives
it a Q of ~10 is +10. I set the SIDE gain to +20. I checked and the ETMX gain is -16 so now
they're at least similar. I have updated the snapshot to reflect the new value.

Hopefully now it will be more well behaved.
  132   Wed Nov 28 16:46:28 2007 ranaConfigurationComputersscientific linux 5.0
I tried installing Scientific Linux on Tiramisu. The installation process was so bad (really)
that I quit after 15 minutes. Its back to booting Ubuntu as if nothing had ever happened. Let
us never speak of Scientific Linux again.
  131   Wed Nov 28 16:18:15 2007 AlbertoMetaphysicsEnvironmentso clean you can eat on it
I tidied up the desks in the lab, brought the Spectrum Analyzers back to the Salumeria (you don't want to know about that), sorted a lot of stuff and boxed up what I didn't know (you can find it in a couple of carton boxes on the table).
The blackmail with the pie might not work next time.
Please, preserve the common sort.


Alberto
  130   Wed Nov 28 12:43:53 2007 AndreyBureaucracy Here was the PDF-file of my presentation

I was making a report with powerpoint presentation during that Wednesday's 40-m meeting.

Here was the pdf-file, but LATER IN THE EVENING I CREATED A WIKI-40M-page describing the algorithm, and now the pdf-file is ON THAT WIKI-40M PAGE.


NOTE ADDED AFTER THE PRESENTATION: I double checked, I am indeed taking the root-mean-square of a difference, as we discussed during my talk.

My slide #17 "Calculation of differential length" was wrong, but now I corrected it.
  129   Wed Nov 28 08:47:29 2007 steveOmnistructureVACrga is out of order
The rga is not working since Nov 10
The controller is broken.
pd65-m-d23
  128   Wed Nov 28 04:21:46 2007 ranaUpdatePSLFSS

Quote:
Rana, Tobin

We looked at the RF PD signal to the FSS (siphoning off a signal via a minicircuits directional coupler) and also took an open loop transfer function of the FSS. In the transfer function we saw the step at 100 kHz (mentioned by Rob) as well as some peculiar behavior at high frequency. The high frequency behavior (with a coupling of ~ -20 dB) turns out to be bogus, as it is still present even with the beam blocked. Rearranging the cabling had no effect; the cause is apparently inside the FSS. The step at 100 kHz turns out to be a saturation effect, as it moved as we lowered the signal amplitude, disappearing as we approached -60 dBm. (Above the step, the measurement data is valid; below, bogus.)

Transfer functions will be attached to this entry.

Some things to check tomorrow: the RF signal to the PC, RF AM generation by the PC, LO drive level into the FSS, RF reflection from the PC, efficiency of FSS optical path, quality of RF cabling.


I would also add to Tobin's entry that we believe what Rob was seeing was saturation.

With the bi-directional coupler in there, the RF signal into the FSS board clearly went UP if moved the offset slider away from zero.
With a scope looking at the IN2 testpoint, we can see that there's less than 2 mV offset at zero slider offset.

One tangential thing we noticed with the coupler is that, in lock, the amount of reflected RF is around the same as that going in to the mixer.
I have always wanted to look at this but have only had uni-directional couplers in the past. I think that the double balanced mixer is inherently
not a 50 Ohm device during the times where the diodes are being switched. IF that's the case we might do better in the future by having an RF
buffer on board just before the mixer to isolate the PD head from these reflections.
  127   Tue Nov 27 20:47:00 2007 tobinUpdatePSLFSS
Rana, Tobin

We looked at the RF PD signal to the FSS (siphoning off a signal via a minicircuits directional coupler) and also took an open loop transfer function of the FSS. In the transfer function we saw the step at 100 kHz (mentioned by Rob) as well as some peculiar behavior at high frequency. The high frequency behavior (with a coupling of ~ -20 dB) turns out to be bogus, as it is still present even with the beam blocked. Rearranging the cabling had no effect; the cause is apparently inside the FSS. The step at 100 kHz turns out to be a saturation effect, as it moved as we lowered the signal amplitude, disappearing as we approached -60 dBm. (Above the step, the measurement data is valid; below, bogus.)

Transfer functions will be attached to this entry.

Some things to check tomorrow: the RF signal to the PC, RF AM generation by the PC, LO drive level into the FSS, RF reflection from the PC, efficiency of FSS optical path, quality of RF cabling.
  126   Tue Nov 27 16:18:58 2007 robConfigurationIOOMC loop
Reduced the common gain to 22dB in the mcup script, so that the WFS would not blow the lock. The above measure of the OLG was done without the mcWFS running, so may be a low estimate as compared to when the alignment is perfect.
  125   Tue Nov 27 15:47:17 2007 robConfigurationIOOMC loop
After the FSS running pretty quick, I checked the MC loop. I used TPA 1&2.

MC loop
UGF: 70kHz
Input Gain: 29dB
Boost Level: 2
phase: 40 deg
  124   Tue Nov 27 15:45:08 2007 robConfigurationPSLFSS loop

It's unclear (to me, at least) what was the end result of the FSS path tweaking before Thanksgiving. Today I measured the open loop gain, and it was still around 100kHz, even with the gain sliders maxed out, but it looked really crappy with a sharp cutoff around the UGF. Then, on a lark, I pushed around the "Input Offset Adjust" slider, which sums an offset into the signal coming out of the mixer. By moving this slider to 7V, I got the UGF to 500kHz with 45 deg of phase. That would be fine, and we could go offset hunting, but the same thing happens if one puts in a large negative value! I don't really understand what's going on, but it seems like weirdness in the electronics. Unfortunately the web interface to the conlog is not running (presumably because the `new' linux1 doesn't have its apache server running) and my command line conlog efforts have been stymied. So, I don't know what the historical settings of this offset are, but zero is definitely not a good setting right now. Here's a snapshot:

FSS
UGF: 500kHz
CG : 24dB
FG : 19dB
input offset: 7V
Phase Adjust: 1.09V
Phase Button: 0
RF Amp Adjust: 7.38V

margins:
phase: 45 deg
gain: 8dB
  122   Mon Nov 26 10:17:31 2007 steveOmnistructureSUSetmy sus damping restored
20 days plot is showing etmy loosing damping 4 times.
I zoomed in with each event. Three of them could of been triggered
by garbage loading just outside. However attachment 2 plot demonstrating that small earthquake or seismic event
did not tripped etmy damping.
The fourth event was preceded by a 4-5 hrs of continous rise of the rms motion at C1:SUS-ETMY_LLPD_VAR
  121   Wed Nov 21 14:31:41 2007 robUpdatePSLFSS twiddle

I `tweaked' the FSS path today. Here's what I did:

1) Shut down the FSS autolocker

2) Turn off FSS servo

3) Assume the beam coming back from the AOM is double-first-order, and don't make any changes large enough to lose it.

4) Tweak the alignment of these components to maximize the incident power on the RC reflected diode:

a) PBS before AOM
b) AOM
c) curved mirror after the AOM

5) Translate the AOM such that the beam moves away from the PZT, then when it levels off (no more power gains with movement),
move it back just a little bit so there's a teensy drop in power. This should but the beam as close to the edge as possible,
but whether or not it's the best place is still to be determined.

6) Lock the FSS, and align the mirrors into the frequency reference cavity.

After all this, the RC transmitted power went from .57 to .73 -- probably not a big enough change to account for the missing loop
gain, but we'll know more once the loop gets measured (after Alberto stops hogging the Agilent network analyzer).

Other possible routes include a systematic check of the upstream path (e.g., the Pockels cell) and just increasing the pickoff fraction for the FSS.
  120   Tue Nov 20 18:35:20 2007 JohnHowToComputersMatLab in Emacs
If you can't get MatLab to run in emacs try adding the following to the .emacs file

(setq matlab-shell-command-switches '("-nojvm"))

This stops the gui opening.

To start MatLab type M-x matlab-shell.
To enter MatLab mode M-x matlab-mode.

I've done this on LINUX3.

To run MatLab in emacs under windows one can use MatLabShell http://www.cs.umb.edu/~ram/matlabShell/index.html