40m QIL Cryo_Lab CTN SUS_Lab CAML OMC_Lab CRIME_Lab FEA ENG_Labs OptContFac Mariner WBEEShop
  40m elog, Page 337 of 357  Not logged in ELOG logo
IDup Date Author Type Category Subject
  16950   Mon Jun 27 13:25:50 2022 CiciUpdateGeneralCharacterizing the Transfer Loop

[Deeksha, Cici]

We first took data of a simple low pass filter, and attempted to perform a fit to both the magnitude and phase in order to find the Z of the components. Once we felt confident in our ability to measure tranfer functions, we took data and plotted the transfer function of the existing control loop of the AUX laser. What we found generally followed the trend of, but was lower than, 10^4/f, which is what we hoped to match, and also had a strange unexplained notch ~1.3 kHz. The magnitude and phase data both got worse after around 40-50 kHz, which we believe is because the laser came out of lock near the end of the run. 

Edit: 

[Attachment 2 and 3] are the frequency response of the low pass filter, curves fitted using least squares in python.

[Attachment 1 and 4] is the same measurement of OLTF of the actual AUX circuit, and the control diagram pointing out the location of excitation and test point.

  16951   Mon Jun 27 13:39:40 2022 DeekshaUpdateElectronicsSetting up the MokuLab

[Cici, Deeksha]

On Friday Cici and I set up the Mokulab to take readings of our loop. The aim is to characterise the PZT, in a similar manner as before, by exciting the circuit using our input noise (a swept sine) and recording the corresponding changes in the output. We used the MokuLab to observe the beat note created by the signals of the AUX and PSL, as well as the ASD of the output signal. The MokuLab simplifies the entire process.

Pictured : The beat note as observed by Cici

  16952   Mon Jun 27 18:54:27 2022 yutaUpdateLSCModulation depths measurement using Yarm cavity scan

[Yehonathan, Yuta]
EDITED by YM on 22:11 June 27, 2022 to correct for a factor of two in the modulation index

Since we have measured optical gain in MICH to be an order of magnitude less compared with Yehonathan's FINESSE model (40m/16923), we measured the power at AS55 RF PD, and measured the modulation depths using Yarm cavity scan.
We found that 50/50 beam splitter which splits AS55 path into RF PD and RF QPD was not included in the FINESSE model. Measured modulation index were as follows:

TEM00 peak height: 0.6226 +/- 0.0237
RF11 peak height: 0.0067 +/- 0.0007
RF55 peak height: 0.0081 +/- 0.0014
RF11 modulation index: 0.208 +/- 0.012
RF55 modulation index: 0.229 +/- 0.020
RF11 modulation index: 0.104 +/- 0.006
RF55 modulation index: 0.114 +/- 0.010

Here, modulation depth m is defined in E=E_0*exp(i*(w*t+m*sin(w_m*t))), and m m/2 equals to square of the intensity ratio between sidebands and TEM00.

Power measurement at AS55 RF PD:
 - ITMY and ITMX single bounce reflection was measured to be 50-60 uW at the front of AS55 RFPD.
 - In the FINESSE model, it was expected to be ~110 uW with 0.8 W input to PRM (0.8 W * 5%(PRM) * 50%(BS) * 50%(BS) * 10%(SRM) * 10%(AS2) gives 100 uW)
 - In AP table, AS55 beam was split into two paths with 50/50 beam splitter, one for AS55 RF PD and one for AS WFS and AS110. This will be included in the FINESSE model.

Modulation depth measurement using Yarm cavity scan:
 - Aligned Yarm using ASS, and unlocked Yarm to get the 2sec scan data of C1:LSC-TRY_OUT_DQ, C1:LSC-POY11_I_ERR_DQ, C1:LSC-AS55_I_ERR_DQ.
 - TRY data was used to get TEM00 peak heights
 - POY11/AS55 data was used to find RF11/RF55 sideband peaks, and height was measured at TRY (see attached).
 - If we define m to be E=E_0*exp(i*(w*t+m*sin(w_m*t))), the amplitude of TEM00 I_00 is proportional to J_0(m) and the amplitude of upper/lower sideband I_f1 is proportional to J_1(m), where J_n(m) is the bessel function of the first kind.
 - m can be calculated using 2*sqrt(I_f1 / I_00).
 - Results were shown above. Error is calculated from the standard deviation of multiple measurements with multiple peaks,
 - The code for doing this lives in https://git.ligo.org/40m/measurements/-/blob/main/LSC/YARM/modulationIndex.ipynb

Discussion:
 - Power at AS55 account for the factor of 2, In the FINESSE model, modulation index of 0.3 was used (could be m=0.3/2 or m=0.3; needs check). These combined can explain a factor of 3 at least (or 6).
 - Gautam's measurement in Jan 2021 (40m/15769) gives almost double modulation index, but I'm not sure what is the definition Gautam used. It agrees with Gautam's measurement in Jan 2021.

  16953   Tue Jun 28 09:03:58 2022 JCUpdateGeneralOrganizing and Cleaning

The plan for the tools in 40m

As of right now, there are 4 tool boxes. X-end, Y-end, Vertex, and the main tool box along the X-arm. The plan is the give each toolbox a set of their own tools. The tools of X-end, Y-end, and Vertex toolboxes will be very similar containing the basic tools such as pliers, screwdrivers, allen ball drivers. Along with this, each tool box will have a tape measure, caliper, level, and other measuring tools we find convinient. 

As for the new toolbox, I have done research and found a few good selections. The only problem I have ran into with this is the width of the tool box corresponding with the prices. The tool cabinet we have now is 41" wide. The issue I have is not in finding another toolbox of the same width, but for a similar price we can find a 54" wide tool cabinet. Would anyone be objected to making a bit more space for this?

How the tools will stay organized.

I the original idea I had was to use a specified color of electrical tape for each tool box. Then to wrap the corresponding tools tools with the same color tape. But it was brought to my attention that the electrical tape would become sticky over time. So, I think the using the label maker would be the best idea. with the labels being 'X' for X-end, 'Y' for Y-end, 'V' for vertex, and 'M' for main toolboxes.

An idea for the optical tables:

Anchal brought it up to me that it is a hassle to go back and forth searching for the correct sizes of Hex Keys and Allen Wrenches. The idea of a pouch on the outside of each optical table was mentioned so I brought this up to Paco. Paco also gave me the idea of a 3D printed stand we could make for allen ball drives. Does anyone have a preference or an idea of what would be the best choice and why? 


A few sidenotes: 

Anchal mentioned to me a while back that there are many cables that are laying on the racks that are not being used. Is there a way we could identify which ones are being used? 

I noticed that when we were vented that a few of the chamber doors were leaning up against the wall and not on a wooden stand like others. Although, the seats for the chamber doors are pretty spacious and do not give us much clearance. For the future ones, could we make something more sleek and put the wider seats at the end chambers?

The cabinets along the Y-Arm are labelled, but do not correspond with all the materials inside or are too full to take in more items. Could I organize these? 
 

  16954   Tue Jun 28 14:24:23 2022 yutaUpdateBHDBHD DC PD signals now also sent to c1lsc to circumvent IPC error

[JC, Yuta]

To circumvent IPC error sending BHD DC PD signals from c1sus2 to c1lsc, DB9 cable from BHD DC PD box sent to c1sus2 is now split and sent also to c1lsc.
They are now available in both

c1sus2 ADC1
C1:X07-MADC1_EPICS_CH16 (DC PD A) and CH17 (DC PD B)

c1lsc ADC1
C1:X04-MADC1_EPICS_CH4 (DC PD A) and CH5 (DC PD B)

Next:
 - Add battery powered SR560 to decouple c1sus2 and c1lsc to avoid the ground loop

  16955   Tue Jun 28 16:26:58 2022 CiciSummaryGeneralVector fitting open loop transfer function/Audio cancellation of optical table enclosure

[Deeksha, Cici]

We attempted to use vectfit to fit our earlier transfer function data, and were generally unsuccessful (see vectfit_firstattempt.png), but are much closer to understanding vectfit than before. Couple of problems to address - finding the right set of initial poles to start with has been very hard, and also however vectfit is plotting the phase data is unwrapping it, which makes it generally unreadable. Still working on how to mess with the vectfit automatically-generated plots. In general, our data is very messy (this is old data of the transfer function from last week), so we took more data today to see if our coherence was the problem (see TFSR785_28-06-2022_161937.pdf). As is visible from the graph, our coherence is terrible, and above 1kHz is almost entirely below 0.5 (or 0.2) on both channels. Figuring out why this is and fixing it is our first priority.

In the process of taking new data, we also found out that the optical table enclosure at the end of the X-arm does a decent job of sound isolation (see enclosure_open.mp4 and enclosure_closed.mp4). The clicking from the shutter is visible on a spectrogram at high frequencies when the enclosure is open, but not when it is closed. We also discovered that the script to toggle the shutter can run indefinitely, which can break the shutter, so we need to fix that problem!

  16956   Tue Jun 28 16:59:35 2022 PacoSummaryALSALS beat allan deviation (XARM)

[Paco]

I took ~ 7 minutes of XALS beatnote data with the XAUX laser locked to the XARM cavity, and the XARM locked to PSL to develop an allan deviation estimator. The resulting timeseries for the channel C1:ALS-BEATX_FINE_PHASE_OUT_HZ_DQ (decimated timeseries in Attachment #1) was turned into an allan variance using the "overlapped variable tau estimator":

\sigma_y^2(n\tau_0, N) = \frac{1}{2n^2\tau_0^2(N - 2n)} \sum_{i=0}^{N-2n-1} (x_{i+2n} - 2x_{i+n} + x_i)^2

Where x_k represents the k-th data point in the raw timeseries, and n\tau_0 are the variable integration intervals under which two point variances are computed (the allan variance is a special case of M-point variance, where M=2). Then, the allan deviation is just the square root of that. Attachment #2 shows the fractional deviation (normalized by the mean beat frequency ~ 3 MHz for this measurement) for 100 integration times spanning the full duration (~ 7 min = 420 s).

The code used for this lives in Git/40m/labutils/measuremens/ALS/


If this estimate is any good, wherever the fractional beatnote deviation reaches a minimum value can be used as a proxy for the longest averaging time that give a statistical increase in SNR. After this timescale, the frequency comparison is usually taken over by "environmental instabilities" which I don't think I can comment further on. In our particular estimate, the 100 second integration gives a fractional deviation of ~ 0.44 %, or absolute deviation of 12.925 kHz.

  16957   Tue Jun 28 17:07:47 2022 AnchalUpdateCalibrationAdded Beatnote channels in demodulation of c1cal

I added today demodulation of C1:LSC-BEATX/Y_FINE_I/Q in the c1cal demodulation where different degrees of freedom can be dithered. For McCal (formerly soCal), we'll dither the arm cavity for which we can use any of the DOFs (like DARM) to send the dither to ETMX/ETMY. Then with green laser locked as well, we'll get the calibration signal from the beatnotes in the demodulaed channels. We can also read right after the mixing in c1cal model and try differnt poles for integration .

I've also added medm screens in the sensing matrix part of LSC screen. These let you see demodulation of beatnote frequency signals.

  16958   Tue Jun 28 18:19:09 2022 TomislavUpdateElectronicsElectronics noise

I measured electronics noise of WFSs and QPD (of the WFS/QPD, whitening, ADC...) by closing PSL and measuring the error signal. It was needed to put the offset in C1:IOO-MC_TRANS_SUMFILT_OFFSET to 14000 cts (without offset the sum of quadrants would give zero, and 14000 cts is the value when the cavity is locked). For WFS that are RF, if there is intensity noise at low frequencies, it is not affecting the measurement.

In the attachment please find the power spectrum of the error signal when the PSL shutter is on and off.

  16959   Tue Jun 28 18:53:16 2022 ranaSummaryALSALS beat allan deviation (XARM)

what's the reasoning behind using df/f_beat instead of df/f_laser ?

Quote:

[Paco]

I took ~ 7 minutes of XALS beatnote data with the XAUX laser locked to the XARM cavity, and the XARM locked to PSL to develop an allan deviation estimator.

 

  16960   Tue Jun 28 22:27:11 2022 YehonathanUpdateBHDMeasurement of input and output electronics noise

{Yuta, Yehonathan}

For MICH noise budgeting we measure the input electronics noise which includes the AS55 RFPD, preamp, demod board, the whitening, and the AA filters, and the ADC noises. To do so we simply close the laser shutter and take the spectrum of C1:LSC-AS55_I_ERR_DQ and C1:LSC-AS55_Q_ERR_DQ shown in attachment 1.

Next, we measured the output electronics noise which includes the DAC, dewhitening and AI filters, and coil driver noises. We disabled the BS watchdog and went to 1X4 rack. We measured the spectrum of one of the lemo outputs on the BS coil driver module using an SR785. Attachment 2 shows the spectrum together with the SR785 dark noise.

  16961   Tue Jun 28 23:10:34 2022 YehonathanUpdateBHDMeasurement of AS55 demod board conversion factor

{Yuta, Yehonathan}

We measured the AS55 demod board conversion from the amplitude of a 55MHz signal to a demodulated signal. We hooked the unused REFL55 LO into the PD input port on the AS55 demod board.

The REFL55 LO was measured to be 1.84 Vpp. The IQ outputs were: I = 0.86 Vpp, Q = 2.03 Vpp giving an amplitude of 2.205 Vpp. The overall conversion factor is sqrt(0.86**2+2.03**2)/(1.82/2)=2.422.

We also set to measure the loss in the RF cable from AS55 PD to the demod board on 1Y2. REFL55 was connected with a long BNC cable to the input of the cable under test. REFL55 at the input was measured to be 1.466 Vpp and 1.28 Vpp at the output signifying a transmission of 87.6%.

  16962   Wed Jun 29 14:28:06 2022 PacoSummaryALSALS beat allan deviation (XARM)

I guess it didn't make sense since f_beat can be arbitrarily moved, but the beat is taken around the PSL freq ~ 281.73 THz. Attachment #1 shows the overlapping tau allan deviation for the exact same dataset but using the python package allantools, where this time I used the PSL freq as the base frequency. This time, I can see the minimum fractional deviation of 1.33e-13 happening at ~ 20 seconds.

Quote:

what's the reasoning behind using df/f_beat instead of df/f_laser ?


Another, more familiar interpretation

The allan variance is related to the beatnote spectral density as a mean-square integral (the deviation is then like the rms) with a sinc window.

\sigma^2_\nu = 2 \int_0^{\infty} S_\nu(f) \lvert \frac{\sin({\pi f \tau})}{\pi f \tau} \lvert ^2 df

  16963   Wed Jun 29 18:53:38 2022 ranaUpdateElectronicsElectronics noise

this is just the CDS error signal, but is not the electronics noise. You have to go into the lab and measure the noise at several points. It can't be done from the control room. You must measure before and afte the whitening.

Quote:

I measured electronics noise of WFSs and QPD (of the WFS/QPD, whitening, ADC...) by closing PSL and measuring the error signal. It was needed to put the offset in C1:IOO-MC_TRANS_SUMFILT_OFFSET to 14000 cts (without offset the sum of quadrants would give zero, and 14000 cts is the value when the cavity is locked). For WFS that are RF, if there is intensity noise at low frequencies, it is not affecting the measurement.

In the attachment please find the power spectrum of the error signal when the PSL shutter is on and off.

 

  16964   Thu Jun 30 17:19:55 2022 DeekshaSummaryElectronicsMeasured Transfer Functions of the Control Loop, Servo (OLTF); got Vectfit working

[Cici, Deeksha]

We were able to greatly improve the quality of our readings by changing the parameters in the config file (particularly increasing the integration and settle cycles, as well as gradually increasing our excitation signals' amplitude). Attached are the readings taken from the same (the files directly printed by ssh'ing the SR785 (apologies)) - Attachment 1 depicts the graph w/ 30 data points and attachment 2 depicts the graph with 300 data points. 

Cici successfully vectfit to the data, as included in Attachment 3. (This is the vectfit of the entire control loop's OLTF). There are two main concerns that need to be looked into, firstly, the manner in which to get the poles and zeros to input into the vectfit program. Similarly, the program works best when the option to enforce stable poles is disabled, once again it may be worth looking into how the program works on a deeper level in order to understand how to proceed. 

Just as the servo's individual transfer function was taken, we also came up with a  plan to measure the PZT's individual transfer function (using the MokuLab). The connections for the same have been made and the Moku is at the Xend (disconnected). We may also have to build a highpass filter (similar to the one whose signal enters the PZT) to facilitate taking readings at high frequencies using the Moku. 

  16965   Thu Jun 30 18:06:22 2022 PacoUpdateALSOptimum ALS recovery - part I

[Paco]

In the morning I took some time to align the AUX beams in the XEND table. Later in the afternoon, I did the same on the YEND table. I then locked the AUX beams to the arm cavities while they were stabilized using POX/POY and turned off the PSL hepa off temporarily (this should be turned on after today's work).

After checking the the temperature slider sign on the spectrum analyzer of the control room I took some out-of-loop measurements of both ALS beatnotes (Attachment #1) by running diaggui /users/Templates/ALS/ALS_outOfLoop_Ref_DQ.xml and by comparing them against their old references (red vs magenta and blue vs cyan); it seems that YAUX is not doing too bad, but XAUX has increased residual noise around and above 100 Hz; perhaps as a result of the ongoing ALS SURF loop investigations? It does look like the OLTF UGF has dropped by half from ~ 11 kHz to ~ 5.5 kHz.

Anyways let this be a reference measurement for current locking tasks, as well as for ongoing SURF projects.

  16966   Thu Jun 30 19:04:55 2022 ranaSummaryPSLPSL HEPA: How what when why

For the PSL HEPA, we wanted it to remain at full speed during the vent, when anyone is working on the PSL, or when there is a lot of dust due to outside conditions or cleaning in the lab.

For NORMAL conditions, the policy is to turn it to 30% for some flow, but low noise.

I think we ought to lock one of the arms on IR PDH and change the HEPA flow settings and plot the arm error signal, and transmitted power for each flow speed to see what's important. Record the times of each setting so that we can make a specgram later

  16967   Thu Jun 30 19:24:24 2022 ranaSummaryPEMeffect of nearby CES construction

For the proposed construction in the NW corner of the CES building (near the 40m BS chamber), they did a simulated construction activity on Wednesday from 12-1.

In the attached image, you can see the effect as seen in our seismometers:

this image is calculated by the 40m summary pages codes that Tega has been shepherding back to life, luckily just in time for this test.

Since our local time PDT = UTC - 7 hours, 1900 UTC = noon local. So most of the disturbance happens from 1130-1200, presumably while they are setting up the heavy equipment. If you look in the summary pages for that day, you can also see the IM lost lock. Unclear if this was due to their work or if it was coincidence. Thoughts?

  16968   Fri Jul 1 08:50:48 2022 yutaSummaryLSCFPMI with REFL/AS55 trial

[Anchal, Paco, Yuta]

We tried to lock FPMI with REFL55 and AS55 this week, but no success yet.
FPMI locks with POX11, POY11 and ASDC for MICH stably, but handing over to 55's couldn't be done yet.

What we did:
 - REFL55: Increased the whitening gain to 24dB. Demodulation phase tuned to minimize MICH signal in I when both arms are locked with POX and POY. REFL55 is noisier than AS55. Demodulation phase and amplitude of the signal seem to drift a lot also. Might need investigation.
 - AS55: Demodulation phase tuned to minimize MICH signal in I when both arms are locked with POX and POY. Whitening gain is 24dB.
 - Script for demodulation phase tuning lives in https://git.ligo.org/40m/scripts/-/blob/main/RFPD/getPhaseAngle.py
 - Locking MICH with REFL55 Q: Kicks BS much and not so stable probably because of noisy REFL55. Offtet also needs to be adjusted to lock MICH to dark fringe.
 - BS coil balancing: When MICH is "locked" with REFL55 Q, TRX drops rapidly and AS fringe gets worse, indicating BS coil balancing is not good. We balanced the coils by dithering POS with different coil output matrix gains to minimize oplev PIT and YAW output manually using LOCKINs.
 - Locking MICH with ASDC: Works nicely. Offset is set to -0.1 in MICH filter and reduced to -0.03 after lock acquisition.
 - ETMX/ETMY actuation balancing: We found that feedback signal to ETMX and ETMY at LSC output is unbalanced when locking with POX and POY. We dithered MC2 at 71 Hz, and checked feedback signals when Xarm/Yarm are locked to find out actuation efficiency imbalance. A gain of 2.9874 is put into C1:LSC-ETMX filter to balance ETMX/ETMY. I think we need to check this factor carefully again.
 - TRX and TRY: We normalized TRX and TRY to give 1 when arms are aligned. Before doing this, we also checked the alignment of TRX and TRY DC PDs (also reduced green scattering for TRY). Together with ETMX/ETMY balancing, this helped making filter gains the same for POX and POY lock to be 0.02 (See, also 40m/16888).
 - Single arm with REFL55/AS55: We checked that single arm locking with both REFL55_I and AS55_Q works. Single arm locking feeding back to MC2 also worked.
 - Handing over to REFL55/AS55: After locking Xarm and Yarm using POX to ETMX and POY to ETMY, MICH is locked with ASDC to BS. Handing over to REFL55_I for CARM using ETMX+ETMY and AS55_Q for DARM using -ETMX+ETMY was not successful. Changing an actuator for CARM to MC2 also didn't work. There might be an unstable point when turning off XARM/YARM filter modules and switching on DARM/CARM filter modules with a ramp time. We also need to re-investigate correct gains and signs for DARM and CARM. (Right now, gains are 0.02 for POX and POY, -0.02 for DARM with AS55_Q (-ETMX+ETMY), -0.02 for CARM with REFL55_I with MC2 are the best we found so far)
 
Next:
 - Measure ETMX and ETMY actuation efficiencies with Xarm/Yarm to balance the output matrix for DARM.
 - Measure optical gains of POX11, POY11, AS55 and REFL55 when FPMI is locked with POX/POY/ASDC to find out correct filter gains for them.
 - Make sure to measure OLTFs when doing above to correct for loop gains.
 - Lock CARM with POY11 to MC2, DARM with POX11 to ETMX. Use input matrix to hand over instead of changing filter modules from XARM/YARM to DARM/CARM.
 - Try using ALS to lock FPMI.

  16969   Fri Jul 1 12:49:52 2022 KojiUpdateIOOMC2 seemed misaligned / fixed

I found the IMC was largely misaligned and was not locking. The WFS feedback signals were saturated and MC2 was still largely misaligned in yaw after resetting the saturation.
It seemed that the MC WFS started to put the large offset at 6:30AM~7:00AM (local).
MC2 was aligned and the lock was recovered then the MC WFS seems working for ~10min now.

  16972   Tue Jul 5 20:05:06 2022 TomislavUpdateElectronicsWhitening electronics noise

For whitening electronics noise for WFS1, I get (attachment). This doesn't seem right, right?

  16973   Wed Jul 6 15:28:18 2022 TegaUpdateSUSOutput matrix diagonalisation : F2P coil balancing

git repo : https://git.ligo.org/40m/scripts/-/tree/main/SUS/OutMatCalc

local dir:  /opt/rtcds/caltech/c1/Git/40m/scripts/SUS/OutMatCalc

 

Here is an update on our recent attempt at diagonalization of the SUS output matrices. There are two different parts to this: the first is coil balancing using existing F2P code which stopped working because of an old-style use of the print function and the second which should now focus on the mixing amongst the various degrees of freedoms (dof) without a DC/AC split I believe. The F2P codes are now working and have been consolidated in the git repo.  

 

TODO:

  • The remaining task is to make it so that we only call a single file that combines the characterization code and filter generation code, preferably with the addition of a safety feature that restores any changed values in case of an error or interruption from the user. The safety functionality is already implemented in the output matrix diagonalization stem of the code, so we just need to copy this over.  
  • Improve the error minimization algorithm for minimizing the cross-coupling between the various dof by adjusting the elements of the output matrix. 

 

Previous work 

https://nodus.ligo.caltech.edu:8081/40m/4762

https://nodus.ligo.caltech.edu:8081/40m/4719

https://nodus.ligo.caltech.edu:8081/40m/4688

https://nodus.ligo.caltech.edu:8081/40m/4682

https://nodus.ligo.caltech.edu:8081/40m/4673

https://nodus.ligo.caltech.edu:8081/40m/4327

https://nodus.ligo.caltech.edu:8081/40m/4326

https://nodus.ligo.caltech.edu:8081/40m/4762

  16974   Wed Jul 6 18:51:20 2022 DeekshaUpdateElectronicsMeasuring the Transfer Function of the PZT

Yesterday, we set up the loop to measure the PZT of the transfer function - the MokuLab sends an excitation (note - a swept sine of 1.0 V) to the PZT. The cavity is locked to the PSL and the AUX is locked to the cavity. In order to measure the effect of our excitation, we take the beat note of the PSL and the AUX. This gives us a transfer function as seen in Attachment 1. The sampling rate of the MokuLab is set to 'ultrafast' (125kHz), so we can expect accurate performance upto 62.5kHz, however, in order to improve our readings beyond this frequency, modifications must be made to the script (MokuPhaseMeterTF) to avoid aliasing of the signal. A script should also be written to obtain and plot the coherence between the excitation and our output. 

Also attached are - Attachment 2 -  the circuit diagram of the setup, and Attachment 3 - the TF data calculated.

Edit - the SR560 as shown in the circuit diagram has since been replaced by a broadband splitter (Minicircuits ZFRSC-42-S+).

  16975   Wed Jul 6 19:58:16 2022 PacoSummaryNoiseBudgetXARM noise budget

[Anchal, Paco, Rana]

We locked the XARM using POX11 and made a noise budget for the single arm displacement; see Attachment #1. The noise budget is rough in that we use simple calibrations to get it going; for example we calibrate the measured error point C1:LSC-XARM_IN1_DQ using the single cavity pole and some dc gain to match the UGF point. The control point C1:LSC-XARM_OUT_DQ is calibrated using the actuator gain measured recently by Yuta. We also overlay an estimate of the seismic motion using C1:PEM-SEIS_BS_X_OUT_DQ (calibrated using a few poles to account for stack and pendulum), and finally the laser frequency noise as proxied by the mode cleaner C1:IOO-MC_F_DQ.

A couple of points are taken with this noise budget, apart from it needing a better calibration;

  1. Overall the inferred residual displacement noise is high, even for our single arm cavity.
    1. By looking at the sim OLTF in foton, it seemed that the single arm cavity loop TF could easily become unstable due to some near-UGF-funkiness likely from FM3 (higher freq boost), so we disabled the automatic triggering on it; the arm stayed locked and we changed the error signal (light blue vs gold (REF1) trace)
  2. The arm cavity is potentially seeing too much noise from the IMC in the 1 to 30 Hz band in the form of laser frequency noise.
    1. Need IMC noise budget to properly debug.
  3. At high frequency (>UGF), there seem to be a bunch of "wiggles" which remain unidentified.
    1. We actually tried to investigate a bit into these features, thinking they might have something to do with misalignment, but we couldn't really find significant correlation.

RXA edit:

  1. we also noticed some weirdness in the calibration of MC_F v. Arm. We think MC_F should be in units of Hz, and Paco calculated the resulting motion as seen by the arm, but there was a factor of several between these two. Need to calibrate MC_F and check. In principle, MC_F will show up directly in ALS_BEATX (with the green PDH lock off), and I assume that one is accurately calibrated. Somehow we should get MC_F, XARM, and ALS_BEAT to all agree. JC is working on calibrating the Mini-Circuits frequency counter, so once that is done we will be in good shape.
  2. we may need to turn on some MC_L feedback for the IMC, so that the MC length follows the NPRO frequency below ~20 Hz.
  3. Need to estimate where the IMC WFS noise is in all of this. Does it limit the MC length stability in any frequency band? How do we determine this?
  4. Also, we want to redo this noise budget today, whilst using AS55 instead of POX. Please measure the Schnupp asymmetry by checking the optimum demod phase in AS55 for locking Xarm v Yarm.
  16976   Wed Jul 6 22:40:03 2022 TegaSummaryCDSUse osem variance to turn off SUS damping instead of coil outputs

I updated the database files for the 7 BHD optics to separate the OSEM variance trigger and the LATCH_OFF trigger operations so that an OSEM variance value exceeding the max of say 200 cnts turns off the damping loop whereas pressing the LATCH_OFF button cuts power to the coil. I restarted the modbusIOC service on c1susaux2 and checked that the new functionality is behaving as expected. So far so good.

 

TODO

Figure out the next layer of watchdogging needed for the BHD optics.  

 

Quote:

[Anchal, JC, Ian, Paco]

We have now fixed all issues with the PD mons of c1susaux2 chassis. The slow channels are now reading same values as the fast channels and there is no arbitrary offset. The binary channels are all working now except for LO2 UL which keeps showing ENABLE OFF. This was an issue earlier on LO1 UR and it magically disappeared and now is on LO2. I think the optical isolators aren't very robust. But anyways, now our watchdog system is fully functional for all BHD suspended optics.

 

  16977   Thu Jul 7 18:18:19 2022 yutaUpdateLSCActuator calibration of ETMX and ETMX

(This is a complete restore of elog 40m/16970 from July 5, 2022 at 14:34)

ETMX and ETMY actuators were calibrated using single arm lock by taking the actuation efficiency ratio between ITMs. Below is the result.

ETMX :  2.65e-9 /f^2 m/counts (0.5007 times ITMX)
ETMY : 10.91e-9 /f^2 m/counts (2.3017 times ITMY)

Motivation:
- ETMX and ETMY actuators seemed to be unbalanced when locking DARM (see 40m/16968)

What we did:
- Reverted to C1:LSC-ETMX_GAIN = 1
- XARM was locked using POX11_I_ERR (42dB whitening gain, 132.95 deg for demod phase) with ETMX and C1:LSC-XARM_GAIN=0.06
- YARM was locked using POY11_I_ERR (18dB whitening gain, -66.00 deg for demod phase) with ETMX and C1:LSC-YARM_GAIN=0.02
- OLTFs for each was measured to be Attachment #1; UGF was ~180 Hz for XARM, ~200 Hz for YARM.
- Measured TF from C1:LSC-(E|I)TM(X|Y)_EXC to C1:LSC-(X|Y)ARM_IN1 (see Attachment #2)
- Took the ratio between ITM actuation and ETM actuation to calculate ETM actuation. For ITM actuation, we used the value measured using MICH (see 40m/16929). The average of the ratio in the frequency range 70-150 Hz was used.

Files:
- Measurement files live in https://git.ligo.org/40m/measurements/-/tree/main/LSC/XARM and YARM
- Script for calculation lives at https://git.ligo.org/40m/scripts/-/blob/main/CAL/ARM/ETMActuatorCalibration.ipynb

Discussion:
- ETMX actuation is 4.12 times less compared with ETMY. This is more or less consistent with what we measured in 40m/16968, but we didn't do loop-correction at that time.
- We should check if this imbalance is as expected or not.

Summary of actuation calibration so far:
BS   : 26.08e-9 /f^2 m/counts (see 40m/16929)
ITMX :  5.29e-9 /f^2 m/counts (see 40m/16929)
ITMY :  4.74e-9 /f^2 m/counts (see 40m/16929)
ETMX :  2.65e-9 /f^2 m/counts (0.5007 times ITMX)
ETMY : 10.91e-9 /f^2 m/counts (2.3017 times ITMY)

 

  16978   Thu Jul 7 18:22:12 2022 yutaUpdateLSCActuator calibration of MC2 using Yarm

(This is also a restore of elog 40m/16971 from Jul 5, 2022 at 17:36)

MC2 actuator calibration was also done using Yarm in the same way as we did in 40m/16970 (now 40m/16977).
The result is the following;
MC2 : -14.17e-9 /f^2 m/counts in arm length (-2.9905 times ITMY)
MC2 :   5.06e-9 /f^2 m/counts in IMC length
MC2 :  1.06e+05 /f^2 Hz/counts in IR laser frequency

What we did:
- Measured TF from C1:LSC-MC2_EXC to C1:LSC-YARM_IN1 during YARM lock using ETMY (see Attachment #1). Note that the sign of MC2 actuation and ITMY actuation is flipped.
- Took the ratio between ITM actuation and MC2 actuation to calculate MC2 actuation. For ITM actuation, we used the value measured using MICH (see 40m/16929). The average of the ratio in the frequency range 70-150 Hz was used (see Attachment #2).
- The actuation efficiency in meters in arm length was converted into meters in IMC length by multiplying it by IMCLength/ArmLength, where IMCLength=13.5 m is half of IMC round-trip length, ArmLength=37.79 m is the arm length.
- The actuation efficiency in meters in arm length was converted into Hz in IR laser frequency by multiplying it by LaserFreq/ArmLength, where LaserFreq=1064 nm / c is the laser frequency.

Files:
- Measurement files live in https://git.ligo.org/40m/measurements/-/tree/main/LSC/YARM
- Script for calculation lives at https://git.ligo.org/40m/scripts/-/blob/main/CAL/ARM/ETMActuatorCalibration.ipynb

Summary of actuation calibration so far:
BS   : 26.08e-9 /f^2 m/counts (see 40m/16929)
ITMX :  5.29e-9 /f^2 m/counts (see
40m/16929)
ITMY :  4.74e-9 /f^2 m/counts (see
40m/16929)
ETMX :  2.65e-9 /f^2 m/counts (0.5007 times ITMX)
ETMY : 10.91e-9 /f^2 m/counts (2.3017 times ITMY)

MC2 : -14.17e-9 /f^2 m/counts in arm length (-2.9905 times ITMY)
MC2 :   5.06e-9 /f^2 m/counts in IMC length

 

NOTE ADDED by YM on July 7, 2022

To account for the gain imbalance in ETMX, ETMY, MC2, LSC violin filter gains were set to:
C1:LSC-ETMX_GAIN = 4.12
C1:LSC-MC2_GAIN = -0.77
This is a temporary solution to make ETMX and MC2 actuation efficiencies from LSC in terms of arm length to be the same as ETMY 10.91e-9 /f^2 m/counts.

I think it is better to make C1:LSC-ETMX_GAIN = 1, and put 4.12 in C1:SUS-ETMX_TO_COIL gains. We need to adjust local damping gains and XARM ASS afterwards.
As for MC2, it is better to put -0.77 in LSC output matrix, since this balancing depends on LSC topology.

  16979   Thu Jul 7 21:25:48 2022 TegaSummaryCDSUse osem variance to turn off SUS damping instead of coil outputs

[Anchal, Tega]

Implemented ramp down of coil bias voltage when the BHD optics watchdog is tripped. Also added a watchdog reset button to the SUS medm screen that turns on damping and ramps up the coil PIT/YAW bias voltages to their nominal values. I believe this concludes the watchdog work.

Quote:

TODO

Figure out the next layer of watchdogging needed for the BHD optics.  

 

  16980   Fri Jul 8 14:03:33 2022 JCHowToVACVacuum Preparation for Power Shutdown

[Koji, JC]

Koji and I have prepared the vacuum system for the power outage on Saturday.

  1. Closed V1 to isolate the main volume.
  2. Closed of VASE, VASV, VABSSCI,VABS, VABSSCO, VAEV, and VAEE.
  3. Closed V6, then close VM3 to isolate RGA
  4. Turn off TP1 (You must check the RPMs on the TP1 Turbo Controller Module)
  5. Close V5
  6. Turn off TP3 (There is no way to check the RPMs, so be patient)
  7. Close V4 (System State changes to 'All pneumatic valves are closed)
  8. Turn off TP2 (There is no way to check the RPMs, so be patient)
  9. Close Vacuum Valves (on TP2 and TP3) which connect to the AUX Pump.
  10. Turn of AUX Pump with the breaker switch wall plug.

From here, we shutdown electronics.

  1. Run /sbin/shutdown -h now on c1vac to shut the host down.
  2. Manually turn off power to electronic modules on the rack.
    • GP316a
    • GP316b
    • Vacuum Acromags
    • PTP3
    • PTP2
    • TP1
    • TP2 (Unplugged)
    • TP3 (Unplugged)

 

  16981   Fri Jul 8 16:18:35 2022 ranaUpdateLSCActuator calibration of MC2 using Yarm

although I know that Yuta knows this, I will just put this here to be clear: the NNN/f^2 calibration is only accurate abouve the pendulum POS eiegenfrequency, so when we estimate the DC part (in diaggui, for example), we have to assume that we have a pendulum with f = 1 Hz and Q ~5, to get the value of DC gain to put into the diaggui Gain field in the calibration tab.

  16982   Fri Jul 8 23:10:04 2022 KojiSummaryGeneralJuly 9th, 2022 Power Outage Prep

The 40m team worked on the power outage preparation. The detailed is summarized on this wiki page. We will still be able to access the wiki page during the power outage as it is hosted some where in Downs.

https://wiki-40m.ligo.caltech.edu/Complete_power_shutdown_2022_07

  16983   Mon Jul 11 11:16:45 2022 JCSummaryElectronicsStartup after Shutdown

[Paco, Yehonathan, JC]

We began starting up all the electronics this morning beginning in the Y-end. After following the steps on the Complete_Power_Shutdown_Procedures on the 40m wiki, we only came across 2 issues.

  1. The Green beam at the Y-End : Turn on the controller and the indicator light began flashing. After waiting until the blinking light becomes constant, turn on the beam. 
  2. C1lsc "could not find operating system"-unable to SSH from Rossa : We found an Elog of how to restart Chiara and this worked. We proceeded by adding this to the procedures of startup.
  16984   Mon Jul 11 11:56:40 2022 YehonathanUpdateBHDMICH AS55 noise budget

I calculated a noise budget for the MICH using AS55 as a sensor. The calculation includes closed-loop TF calculations.

The notebook and associated files can be found on https://git.ligo.org/40m/bhd/-/blob/master/controls/compute_MICH_noisebudget.ipynb.

Attachment 1 shows the loop diagram I was using. The equation describing the steady-state of the loop is

\left[\mathbb{I}-G \right]\begin{pmatrix} \gamma \\ \delta \\ \Delta\end{pmatrix} = \begin{pmatrix} \alpha \\ \beta \\ \epsilon\end{pmatrix}

, where G is the adjacency matrix given by

G=\begin{pmatrix} 0 & 0 & AE_2\\ 0 & 0 & BE_2 \\ E_1C & E_1D & 0 \end{pmatrix}

First, the adjacency matrix G is constructed by stitching the small ABCDE matrices together. Once the inverse of (I-G) is calculated we can simply propagate any noise source to \delta and then calculate \left[\mathbb{I}-E(CA+DB)\right]B^{-1}\delta to estimate the displacement of the optics. 

Attachment 2 shows the calculated noise budget together with Yuta's measurement.

All the input and output electronics are clumped together for now. Laser noise is irrelevant as this is a heterodyne measurement at 55MHz.

It seems like there is some mismatch in the calibration of the optical gain between the measurement and model. The missing noise at 3-30Hz could be due to angle-to-length coupling which I haven't included in the model.

  16985   Mon Jul 11 15:26:12 2022 JCHowToVACStartup after Power Outage

[Koji, Jc]

Koji and I began starting that Vacuum system up.

  1. Reverse order step 2 of shutting down electronics. Anthing after, turn on manually.
  2. If C1vac does not come back, then restart by holding the reset button.
  3. Open VA6
  4. Open VASE, VASV,VABSSCI, VABS, VABSSCO, VAEV, and VAEE
  5. Open V7
  6. Check P3 and P2, if they are at high pressure, approx. 1 Torr range, then you must use the roughing pumps.
  7. Connect Rotary pump tube. (Manually)
  8. Turn on AUX Pump
  9. Manually open TP2 and TP3 valves.
  10. Turn on TP2 and TP3, when the pumps finish startup, turn off Standby to bring to nominal speed.
  11. Turn on RP1 and RP3
  12. Open V6
  13. Once P3 reaches <<1 Torr, close V6 to isolate the Roughing pumps.
  14. When TP2 and TP3 are at nominal speed, open V5 and V4.
  15. Now TP1 is well backed, turn on TP1.
  16. When TP1 is at nominal speed, Open V1.
  16986   Mon Jul 11 17:25:43 2022 TegaUpdateVACfixed obsolete reference bug in serial_XGS600 service

Koji noticed that the FRG sensors were not updating due to reference to an obsolete modbusIOC_XGS service, which was used temporarily to test the operation of the serial XGS sensor readout to EPICS. The information in this service was later moved into modbusIOC.service but the dependence on the modbusIOC_XGS.service was not removed from the serial_XGS600.service. This did not present any issue before the shutdown, probably bcos the obsolete service was already loaded but after the restart of c1vac, the obsolete service file modbusIOC_XGS.service was no longer available. This resulted in  serial_XGS600.service throwing a failure to load error for the missing obsolete modbusIOC_XGS service. The fix involved replacing two references to 'modbusIOC_XGS' with 'modbusIOC' in  /opt/target/services/serial_XGS600.service.

I also noticed that the date logged in the commit message was Oct 2010 and that I could not do a push from c1vac due to an error in resolving git.ligo.org. I was able to push the commit from my laptop git repo but was unable to do a pull on c1vac to keep it synced with the remote repo.

  16987   Mon Jul 11 17:41:52 2022 KojiHowToVACStartup after Power Outage

- Once the FRG gauge readings are back (see next elog by Tega), I could open V1 to pump down the main vacuum manifold.
- TP2/TP3 were brought back to stand-by mode (slower spinning)
- V7 was closed to separate the annuli side and TP1

During the vacuum recovery, I saw TPs were automatically turned on as soon as the backing pumps were engaged. I could not figure out what caused this automation.

Also, I saw some gate valve states changed while I was not touching them. e.g. V7 was close / VM3 was open / etc
I really had no idea what/who was handling these.

As of ~18:00 local, the main volume pressure is ~2e-5 torr and ready to open the PSL shutter.

  16988   Mon Jul 11 19:29:23 2022 PacoSummaryGeneralFinalizing recovery -- timing issues, cds, MC1

[Yuta, Koji, Paco]

Restarting CDS

We were having some trouble restarting all the models on the FEs. The error was the famous 0x4000 DC error, which has to do with time de-synchronization between fb1 and a given FE. We tried a combination of things haphazardly, such as reloading the gpstime process using

controls@fb1:~ 0$ sudo systemctl stop daqd_*
controls@fb1
:~ 0$ sudo modprobe -r gpstime
controls@fb1:~ 0$ sudo modprobe gpstime
controls@fb1:~ 0$ sudo systemctl start daqd_*
controls@fb1:~ 0$ sudo systemctl restart open-mx.service

without much success, even when doing this again after hard rebooting FE + IO chassis combinations around the lab. Koji prompted us to check the local times as reported by the gpstime module, and comparing it to network reported times we saw the expected offset of ~ 3.5 s. On a given FE ("c1***") and fb1 separately, we ran:

controls@c1***:~ 0$ timedatectl
  Local time: Mon 2022-07-11 16:22:39 PDT
  Universal time: Tue 2022-07-11 23:22:39 UTC
       Time zone: America/Los_Angeles (PDT, -0700)
       NTP enabled: yes
       NTP synchronized: no
 RTC in local TZ: no
       DST active: yes
 Last DST change: DST began at
                  Sun 2022-03-13 01:59:59 PST
                  Sun 2022-03-13 03:00:00 PDT
 Next DST change: DST ends (the clock jumps one hour backwards) at
                  Sun 2022-11-06 01:59:59 PDT
                  Sun 2022-11-06 01:00:00 PST
controls@fb1:~ 0$ ntpq -p
     remote           refid      st t when poll reach   delay   offset  jitter
==============================================================================
 192.168.123.255 .BCST.          16 u    -   64    0    0.000    0.000   0.000

which meant a couple of things:

  1. fb1 was serving its time (broadcast to local (martian) network)
  2. fb1 was not getting its time from the internet
  3. c1*** was not synchronized even though fb1 was serving the time

By looking at previous elogs with similar issues, we tried two things;

  1. First, from the FEs, run sudo systemctl restart systemd-timesyncd to get the FE in sync; this didn't immediately solve anything.
  2. Then, from fb1, we tried pinging google.com and failed! The fb1 was not connected to the internet!!!

We tried rebooting fb1 to see if it connected, but eventually what solved this was restarting the bind9 service on chiara! Now we could ping google, and saw this output

controls@fb1:~ 0$ ntpq -p
     remote           refid      st t when poll reach   delay   offset  jitter
==============================================================================
+tor.viarouge.ne 85.199.214.102   2 u  244 1024  377  144.478    0.761   0.566
*ntp.exact-time. .GPS.            1 u   93 1024  377  174.450   -1.741   0.613
 time.nullrouten .STEP.          16 u    - 1024    0    0.000    0.000   0.000
+ntp.as43588.net 129.6.15.28      2 u  39m 1024  314  189.152    4.244   0.733
 192.168.123.255 .BCST.          16 u    -   64    0    0.000    0.000   0.000 

meaning fb1 was getting its time served. Going back to the FEs, we still couldn't see the ntp synchronized flag up, but it just took time after a few minutes we saw the FEs in sync! This also meant that we could finally restart all FE models, which we successfully did following the script described in the wiki. Then we had to reload the modbusIOC service in all the slow machines (sometimes this required us to call sudo systemctl daemon-reload) and performed burt restore to a last Friday's snap file collection.


IMC realign and MC1 glitch?

With Koji's help PMC locked, and then Yuta and Paco manually increased the input power to the IFO by rotating the waveplate picomotor to 37.0 deg. After this, we noticed that the MC REFL spot was not hitting the camera, so maybe MC1 was misaligned. Paco checked the AP table and saw the spot horizontally misaligned on the camera, which gave us the initial YAW correction on MC1. After some IMC recovery, we saw only MC1 got spontaneously kicked along both PIT and YAW, making our alignment futile. Though not hard to recover, we wondered why this happened.

We went into the 1X4 rack and pushed MC1 suspension cables in to disregard loose connections, but as we came back into the control room we again saw it being kicked randomly! We even turned damping off for a little while and this random kicking didn't stop. There was no significant seismic motion at the time so it is still unclear of what is happening.

  16989   Tue Jul 12 09:14:50 2022 ranaUpdateBHDMICH AS55 noise budget

Looking good:

  • I think the notches you see in he measured noise are a clue as to the excess noise source. You can try turning some notches on/off.
  • Laser noise does matter a bit more subtley: the low freq noise couples to AS55 through the RMS deviation of the MICH loop from the zero crossing, and the noise of the 55 MHz modulation.
  • Jitter in the IMC couples to MICH through the misalignment of the Michelson.
  • As you rightly note, the optical lever feedback on the ITMs and BS also make length noise through the suspension actuator imbalance and the spot mis-centering.
  16990   Tue Jul 12 09:25:09 2022 ranaUpdateIOOIMC WFS

MC WFS Demod board needs some attention.

Tomislav has been measuring a very high noise level in the MC WFS demod output (which he promised to elog today!). I thought this was a bogus measurement, but when he, and Paco and I tried to measure the MC WFS sensing matrix, we noticed that there is no response in any WFS, although there are beams on the WFS heads. There is a large response in MC2 TRANS QPD, so we know that there is real motion.

I suspect that the demod board needs to be reset somehow. Maybe the PLL is unlocked or some cable is wonky. Hopefully not both demod boards are fried.

Please leave the WFS loops off until demod board has been assessed.

  16991   Tue Jul 12 13:59:12 2022 ranaSummaryComputersprocess monitoring: Monit

I've installed Monit on megatron and nodus just now, and will set it up to monitor some of our common processes. I'm hoping that it can give us a nice web view of what's running where in the Martian network.

  16992   Tue Jul 12 14:56:17 2022 TomislavSummaryElectronicsElectronics noise measurements

[Paco, Tomislav]

We measured the electronics noise of the demodulation board, whitening board, and ADC for WFSs, and OPLEV board and ADC for DC QPD in MC2 transmission. We were using SR785.

Regarding the demodulation board, we did 2 series of measurements. For the first series of measurements, we were blocking WFS (attachment 1) and measuring noise at the output of the demod board (attachment 2a). This measurement includes dark noise of the WFS, electronics noise of demod board, and phase noise from LO. For the second series of the measurements, we were unplugging input to the demod board (attachment 2b & 2c is how they looked like before unplugging) (the mistake we made here is not putting 50-ohm terminator) and again measuring at the output of the demod board. This measurement doesn't include the dark noise of the WFS. We were measuring it for all 8 segments (I1, I2, I3, I4, Q1, Q2, Q3, Q4). The dark noise contribution is negligible with respect to demod board noise. In attachments 3 & 4 please find plots that include detection and demodulation contributions for both WFSs.

For whitening board electronics noise measurement, we were terminating the inputs (attachment 5) and measuring the outputs (attachment 6). Electronics noise of the whitening board is in the attachments 7 & 8.

For ADC electronics noise we terminated ADC input and measured noise using diaggui (attachments 9 & 10). Please find these spectra for WFS1, WFS2, and MC TRANS in attachments 11, 12 & 13.

For MC2 TRANS we measured OPLEV board noise. We did two sets of measurements, as for demod board of WFSs (with and without QPD dark noise) (attachments 14, 15 & 16). In the case of OPLEV board noise without dark noise, we were terminating the OPLEV input. Please find the electronics noise of OPLEV's segment 1 (including dark noise which is again much smaller with respect to the OPLEV's electronics noise) in attachment 17.

For the transfer functions, demod board has flat tf, whitening board tf please find in attachment 18, ADC tf is flat and it is (2**16 - 1)/20 [cts/V], and dewhitening tf please find in attachment 19. Also please find the ASD of the spectral analyzer noise (attachment_20).

Measurements for WFS1 demod and whitening were done on 5th of July between 15h and 18h local time. Measurements for WFS2 demod and whitening were done on 6th of July between 15h and 17h local time. All the rest were done on July 7th between 14h and 19h. In attachment 21 also find the comparison between electronics noise for WFSs and cds error signal (taken on the 28th of June between 17h and 18h). Sorry for bad quality of some pictures.

  16993   Tue Jul 12 18:35:31 2022 Cici HannaSummaryGeneralFinding Zeros/Poles With Vectfit

Am still working on using vectfit to find my zeros/poles of a transfer function - now have a more specific project in mind, which is to have a Red Pitaya use the zero/pole data of the transfer function to find the UGF, so we can check what the UGF is at any given time and plot it as a function of time to see if it drifts (hopefully it doesn't). Wrestled with vectfit more on matlab, found out I was converting from dB's incorrectly (should be 10^(dB/20)....) Intend to read a bit of a book by Bendat and Piersol to learn a bit more about how I should be weighting my vectfit. May also check out an algorithm called AAA for fitting instead.

  16994   Tue Jul 12 19:46:54 2022 PacoSummaryALSHow (not) to take NPRO PZT transfer function

[Paco, Deeksha, rana]

Quick elog for this evening:

  • Rana disabled MC servo .
  • Slow loop also got disengaged.
  • AUX PSL beatnote is best taken with *free running lasers* since their relative frequency fluctuations are lowest than when locked to cavities.
  • DFD may be better to get PZT transfer funcs, or get higher bandwidth phase meter.
  • Multi instrument to be done with updated moku
  • Deeksha will take care of updated moku
  16995   Wed Jul 13 07:16:48 2022 JCUpdateElectronicsChecking Sorensen Power Supplies

[JC]

I went around 40m picking up any Sorensens that were laying around to test if they worked, or in need of repair. I gathered up a total of 7 Sorensens and each one with a Voltmeter. I made sure the voltage would rise on the Sorenson as well as the voltmeter, maxing out at ~33.4 Volts. For the current, the voltmeter can only rise to 10 Amps before it is fused. Many of the Sorensons that I found did not have their own wall connection, so I had to use the same one for multiple.

From these 7, I have found 5 that are well. One Sorenson I have tested has a output shortage above 20V and the other has yet to be tested.

  16996   Wed Jul 13 10:54:39 2022 YehonathanUpdateBHDMICH AS55 noise budget

I fixed some mistakes in the budget:

1. The BS pendulum resonance was corrected from 0.8Hz to 1Hz

2. Added missing X3 filter in the coil filters

3. Optical gain is now computed from MICH to AS55 instead of BS to AS55 and is calculated to be: 9.95e8 cts/m.

4. Coil driver gain is still unmeasured but it is found to be 1.333 to make the actuation calibration from BS to MICH match the measurement (see attachment 1).

Attachment 2 shows the resulting MICH OLTF.

Laser noise was added to the budget in a slightly ad-hoc fashion (will fix later): Yuta and I measured MC_F and computed MC_F*(Schnupp asymmetry)/(Laser frequency). Attachment 3 shows the updated noise budget.

  16997   Wed Jul 13 12:49:25 2022 PacoSummarySUSSUS frozen

[Paco, JC, Yuta]

This morning, while investigating the source of a burning smell, we turned off the c1SUS 1X4 power strip powering the sorensens. After this, we noticed the MC1 refl was not on the camera, and in general other vertex SUS were misaligned even though JC had aligned the IFO in the morning to almost optimum arm cavity flashing. After a c1susaux modbusIOC service restart and burt restore, the problem persisted.

We started to debug the sus rack chain for PRM since the oplev beam was still near its alignment so we could use it as a sensor. The first weird thing we noticed was that no matter how much we "kicked" PRM, we wouldn't see any motion on the oplev. We repeatedly kicked UL coil and looked at the coil driver inputs and outputs, and also verified the eurocard had DC power on which it did. Somehow disconnecting the acromag inputs didn't affect the medm screen values, so that made us suspicious that something was weird with these ADCs.


Because all the slow channels were in a frozen state, we tried restarting c1susaux and the acromag chassis and this fixed the issue.

  16998   Wed Jul 13 13:26:44 2022 ranaSummaryElectronicsElectronics noise measurements

as I said to you yesterday, I don't think image 2a shows the output of the demod board. The output of the demod board is actually the output connector ON the demod board. What you are showing in 2a, is the signal that goes from the whitening board to the ADC I believe. I may be msitaken, so please check with Tega for the signal chain.

  16999   Wed Jul 13 13:30:48 2022 YehonathanUpdateBHDadd Laser RIN to MICH budget

the main laser noise coupling for a Michelson is because of the RIN, not the frequency noise. You can measure the RIN, in MC trans or at the AS port by getting a single bounce beam from a single ITM.

  17000   Wed Jul 13 17:30:19 2022 KojiUpdateCDSToo huge script_archive

I wanted to check the script archive to see some old settings. I found that the script archive inflated to huge volume (~1TB).
The size of the common NFS volume (/cvs/cds) is 3TB. So it is really significant.

- The scripts living in /opt/rtcds/caltech/c1/scripts are archived daily in /cvs/cds/caltech/scripts_archive as bz2 files. This is done by crontab of megatron (see https://wiki-40m.ligo.caltech.edu/Computers_and_Scripts/CRON)

- In fact, the script folder (say old script folder) /opt/rtcds/caltech/c1/scripts has the size of 10GB. And we have the compressed copy of thi s everyday.

- This large script folder is due to a couple of huge files/folders listed below

  • (scripts)/MEDMtab is 5.3GB / This is probably related to the web MEDM view (on nodus) but I don't think the web page is not updated. (i.e. the images are unused)
  • (scripts)/MC/logs/AutoLocker.log 2.9GB / This is just the accumulated MC autolocker log.
  • (scripts)/GigE 780M / This does not look like scripts but source and object files
  • (scripts)/Admin/n2Check.log 224M / This is important but increases every minute.
  • (scripts)/ZI 316MB / Zurich Instrument installation. This should not be here.

Here I propose some changes.
For the script archive

  • We can remove most of the scripts for the past (say ~2019). We leave an archive file per month.
  • For the scripts in 2020, we leave a weekly archive.
  • For 2021 and 2022, we leave all the archive files.

For the existing large files/folders

  • MEDMtab: the stored files are redundant with the burt snapshots. Remove the image files. Also, we want to move the image-saving location.
  • Autolocker.log: simply zap it
  • n2Check.log: we should move the saving location
  • GigE /ZI: they need a new home where the daily copy is not taken.
  17001   Wed Jul 13 18:58:17 2022 KojiUpdateIOOIMC suspecion

This is just my intuition but the IMC servo seems not so optimized. I can increase the servo gain by 6~10dB easily. And I couldn't see that the PC drive went mad (red) as I increase the gain (=UGF).
The IMC needs careful OLTF measurements as well as the high freq spectrum observation.

It seems that I have worked on the IMC servo tuning in 2014 July/Aug. Checking these elogs would be helpful.