  8998   Mon Aug 12 15:39:40 2013 JenneUpdateASCPRCL ASS screens in place

I have added the PRCL ASS to the main ASS screen, and created the servo and lockin screens.  The filters loaded are the same as those used for the arms (bandpasses and lowpasses for the lockins, and an integrator for the servo).

I'm going to try to lock, and get the ASS to work.

  8999   Mon Aug 12 17:30:03 2013 KojiUpdateASCPRCL ASS software in place

Why POPDC???

  9001   Mon Aug 12 23:13:14 2013 JenneUpdateASCPRCL ASS software in place

Quote:

Why POPDC???

 I guess I was thinking that POPDC was a proxy for any type of PRCL lock.  Even if we're sideband locked, there is still some signal in POPDC (although it is very small relative to a carrier lock - ~40cts vs. 1,000cts).  However, as soon as this question was asked of me, I realized that one of the 2f demodulated signals made more sense. 

Since I want the ability to choose between POP110 and POP22, I have put a little 1x3 input matrix before the PRCL lockins in the ASS model.  Since POPDC was already there, I included it as an option in the matrix (in case we ever want to do some PRCL ASS after we have some carrier resonating as well). 

  9126   Thu Sep 12 01:06:09 2013 JenneUpdateASCSRCL ASS implemented

I have modified the ASS model to also have an ASS for SRCL.  The input options are POPDC, POP110, AS110.  I suppose I could/should have included ASDC.

Screens are modified / made.  I haven't finished setting the servo gains and oscillator amplitudes, and all that jazz yet. 

Using the parameters that Koji had in elog 9116, I was able to get nice long DRMI locks (several on a ~10 minute time scale). 

I tried some pseudo-ANDing for the triggers, to no avail.  I was trying to have the trigger matrix row for the SRCL loop have 1*POP22 and 0.02*AS110, where the 0.02 is to scale AS110 so that it has a similar amplitude to POP22.  I then set threshold levels to ~250 for up, and 100 for down (I tried several different values for the up threshold).  I was watching the TRIG_MON_FAST channels for both PRCL and SRCL, and I wasn't able to get SRCL to be triggered only at the same times as PRCL using this technique.  Since we can get the DRMI to lock, perhaps my AND logic for the triggers is a low priority, but I think we'll need something like that if we want real logic.

 

  9207   Sun Oct 6 20:55:08 2013 ranaSummaryASCMC WFS Limits set based on 40 days of trends

MC3 watchdog gets tripped sometimes when lock is lost. I noticed that there were no limits set in the MC WFS drive. The attached plot shows that over 40 days, the OUT16 channels from the WFS don't exceed 1000 counts. So I've set the limit to be 2000 in all 6 of the MC ASCPIT/YAW filter banks. Please don't turn them off.

OUT16 is really not the right way to measure this, but for some reason, we don't have any DQ channels from the MC WFS screen ??? So we're not able to measure the trend of the high frequency drive signal.

So I added the WFS(1,2)_I_(PIT,YAW)_OUT_DQ and WFS(1,2)_(PIT,YAW)_OUT_DQ channels to the c1ioo.mdl at 2048 Hz. I used Jamie's excellent 'rtcds' utility to build and install:

1) after making the edits to c1ioo.mdl I saved the file/

2) sshing to c1ioo

3) rtcds stop c1ioo

4) rtcds make c1ioo

5) rtcds install c1ioo

6) rtcds start c1ioo

7) telnet fb 8087

8) daqd> shutdown

That seemed to do it OK.

Unfortunately, all of the instructions that we have in the Wiki for adding channels and model building are misleading and don't mention any of this. There are a few different methods listed which all instruct us to do the whole make and make install business in a bunch of non existent directories.

Attachment 1: mcwrfs_trend.png
mcwrfs_trend.png
  9249   Thu Oct 17 13:26:13 2013 JenneUpdateASCPOP QPD realigned

I locked the PRMI, and tried to turn on the ASS, but this caused PRMI to lose lock. 

Since this is similar to what happened the other night (see elog 9243, 2nd big paragraph), I looked into it a little further.  I noticed that the POP QPD pitch was very close to the edge of the QPD, so I went out and (while PRMI was locked) recentered the POP QPD.  After doing so, I was able to run the PRM ASS, and it worked very nicely, just as it has before.  So, it looks like something drifted, such that the optimal PRM alignment caused the POP beam to not be fully on the QPD.  Since the ASC loop is triggered by PRMI lock, and is constantly on, falling off the QPD causes lockloss.

While I was out there, I tweaked up the PMC pitch alignment yet again.  The FSS numbers all looked reasonable, however PMC transmission was ~0.75 .  I did a tiny bit of work in pitch, and now we're back to 0.83 transmission. 

  9460   Thu Dec 12 21:30:52 2013 JenneUpdateASCPRMI-relevant oplevs centered

The ITM oplevs were pretty close to the edge of their ranges, and none of the oplevs have been centered in a while, so I centered ITMX, ITMY, BS, PRM after having done alignment (arms, then PRMI).

  9469   Fri Dec 13 19:33:56 2013 DenUpdateASCETM X,Y QPDs

I have modified/compiled/installed/restarted c1scx and c1scy models to include arm transmission QPDs in angular controls.

For initial test I have wired normalized QPD pitch and yaw outputs to ASC input of ETMs. This was done to keep the signals inside the model.

QPD signals are summed with ASS dither lines and control. So do not forget to turn off QPD output before turning on dither alignment.

Medm screens were made and put to medm/c1sc{x,y}/master directory. Access from sitemap is QPDs -> ETM{ X,Y} QPD

  9558   Wed Jan 15 18:42:57 2014 JenneUpdateASCPOP ASC QPD offline for a few hours this afternoon

I was in the lab, near the south end of the ITMX oplev table, looking for something, and I bumped the POP ASC QPD's power supply.  I thought that it was fine, but did not adequately check it.  When EricQ asked me just now about why the PRC is so wobbly today, I checked, and the power for the QPD wasn't properly connected (it's kind of a crappy connector, that if you nudge, contacts or loses contact).  Anyhow, I restored power to the QPD, and the PRC looks a little more stable now.  My fault for not checking more carefully, and my apologies to Q and Gabriele for their frustrations this afternoon.

  9619   Mon Feb 10 18:59:25 2014 JenneUpdateASCPRM ASC better, but not great yet

I have turned off the 3.2Hz res gains in the PRC ASC loops, since those seem to make the loops unstable. 

Right now the pitch gain is -0.001, with FM1,3,9 on.  Yaw gain is -0.004, with FM1,3,9 on. 

Pitch gain can't increase by factor of 2 without oscillating. 

I tried to take transfer functions, but I think the ASC situation is really confusing, since I have OSEM damping, oplev damping, and this POP QPD damping on the PRM.  It's hard to get coherence without knocking the PRC out of lock, and it keeps looking like my gain is 0dB, with a phase of 0 degrees, from ~1 Hz to ~10 Hz.  Outside that range I haven't gotten any coherence.  Moral of the story is, I'm kind of puzzled. 

Anyhow, as it is right now, the ASC helps a bit, but not a whole lot.  I increased the trigger ON value, so that it shouldn't kick the PRM so much.  I wish that I had implemented a delay in the trigger, but I'm not in the mood to mess with the simulink diagrams right now.

  9620   Mon Feb 10 19:56:10 2014 ranaUpdateASCPRM ASC better, but not great yet

Ignoring the OSEM damping loops, the oplev servo loops make it so that the POP ASC loops do not see a simple pendulum plant, but instead see the closed loop response. Since the filter in the OL bank is proportional to f, this means that the open loop gain (OLG):

 

Which means that the CLG that the ASC sees is going to dip below unity in the band where the OL is on. For example, if the OL loop has a UGF of 5 Hz, it also has a lower UGF of ~0.15 Hz, which means that the ASC needs to know about this modified plant in this band.

For i/eLIGO, we dealt with this in this way: anti-OL in iLIGO

  9769   Mon Mar 31 23:57:22 2014 KojiSummaryASCPRM ASC characterization / design

A series of measurements / calculations for the PRM ASC characterization and servo design

1) Actuator characterization

The actuator responses of the PRM in pitch and yaw were measured (attachment figure 1). I believed the calibration of the oplev QPD to be
1 count/urad. The oplev servo loops were turned off at the FM inputs, and the filter banks were turned off so that the response has the open
loop transfer function except for the servo filter.

The measured transfer functions were fitted with LISO. The LISO results (c.f. the source codes) were shown in the figure. The responses also
include the 60Hz comb filter present in the input filters. The responses are well approximated by the single pendulum with f0 of 0.6-0.8 and q of 3.5 and 6.3.

From this measurement, the actuator responses of the PRM at DC are estimated to be 2.2 urad/cnt and 1.8 urad/cnt in pitch and yaw, respectively.

2) Sensor response of the POP QPD

As we already know how the actuators respond, the QPD optical gain can be characterized by measuring the actuator response of the QPD
(attachment figure 2). The QPD signals are such noisy that the response above 1Hz can't be measured with sufficient coherence. Below 1Hz,
the response is well represented by the actuator response measured with the oplev. From this measurement, the optical gains of the QPD
with respect to the PRM motion are 650 cnt/urad and 350 cnt/urad.

3) Open loop transfer function of the current ASC servo

By combining the above information with the servo setting of the servo filters, the open loop transfer functions of the PRM QPD ASC loops
were estimated (attachment figure 3). Actually the expected suppression of the fluctuation is poor. The yaw loop seems to have
too low gain, but in fact increasing gain is not so beneficial as there is no reasonable phase margin at higher frequency.

With the estimated openloop transfer functions and the measured free-running angular fluctuation, the suppressed angular spectra can be
estimated (attachment figure 4). This tells us that the suppression of the angular noise at around 3Hz is not sufficient in both pitch and yaw.
As there is no mechanical resonance in the actuator response at the frequency, intentional placement of poles and zeros in the servo filter is necessary.

4) Newly designed ASC filter

Here is the new design of the QPD ASC servo (attachment figure 5). The target upper UGF is 10Hz with the phase margin of 50 to 60deg.
The servo is AC coupled so that we still can tweak the alignment of the mirror.

As this servo is conditionally stable, at first we should close the loops with stable filter and then some boosts should be turned on.
Estimated suppressed fluctuation is shown in the attachment figure 6. We can see that the fluctuation was made well white between 0.5Hz to 10Hz.

The filter design is shown as follows:


Pitch
FM1: zero at 0Hz, pole at 2000Hz, gain at 2000Hz = 2000

FM3: (boost)
zero: f: 0.5Hz q: 1  /  4.5Hz, q: 1 / f: 1Hz, q: 3
pole: f: 2Hz q: 3  / f: 2.7Hz, q: 2  / f: 1Hz, q: 15

FM9: (HF Roll-off)
pole: f: 40Hz q: 1.7
 
Servo gain: -0.028

Yaw
FM1: zero at 0Hz, pole at 2000Hz, gain at 2000Hz = 2000

FM3: (boost)
zero: f: 0.7Hz q: 2  /  3Hz, q: 7 / f: 2Hz, q: 6
pole: f: 1.02Hz q: 10  / f: 4.5Hz, q: 0.8  / f: 1.5Hz, q: 10

FM9: (HF Roll-off)
pole: f: 40Hz q: 1.7
 
Servo gain: -0.0132


 

Attachment 1: PRM_OPLEV.pdf
PRM_OPLEV.pdf
Attachment 2: PRM_QPD.pdf
PRM_QPD.pdf
Attachment 3: OLTF_design.pdf
OLTF_design.pdf
Attachment 4: QPD_spe.pdf
QPD_spe.pdf
Attachment 5: OLTF_design2.pdf
OLTF_design2.pdf
Attachment 6: QPD_spe2.pdf
QPD_spe2.pdf
Attachment 7: 140328.zip
  9773   Tue Apr 1 22:03:44 2014 KojiSummaryASCNew PRM ASC is running

[Koji Jenne]

New PRM ASC was implemented. [to be cnt'd]

  9777   Wed Apr 2 19:50:12 2014 KojiSummaryASCNew PRM ASC is running

As the designed ASC filters in this entry had too little phase margins (~10deg), I had to compromise the servo design.

The design was modified and tested again. This will be reported by a following entry.

Incidentally, I have adjusted the demodulation phases of REFL33/55/165 for PRMIsb so that the PRCL is eliminated from the Q signals.

REFL33    125.5 deg -> +136.5 deg
REFL55      45.0 deg -> +  25.0 deg
REFL165   -79.5 deg -> +  44.5 deg

This change of the demod phase for REFL165 was a bit surprising.
I did not check the sign, so it could be -135.5 deg. But still this is a bit change.

  9779   Wed Apr 2 23:08:51 2014 KojiSummaryASCNew PRM ASC is running

The new PRM ASC design


Pitch
FM1: zero at 0Hz, pole at 2000Hz, gain at 2000Hz = 2000

FM5: (boost)
zero: f: 0.5Hz q: 1  /  4Hz, q: 2 / f: 1Hz, q: 3
pole: f: 2Hz q: 3  / f: 2.7Hz, q: 2  / f: 1Hz, q: 15

FM9: (HF Roll-off)
pole: f: 40Hz q: 1/Sqrt(2) (2nd order butterworth)

Servo gain: -0.023

Yaw
FM1: zero at 0Hz, pole at 2000Hz, gain at 2000Hz = 2000

FM5: (boost)
zero: f: 0.5Hz q: 1  /  4Hz, q: 2 / f: 1.5Hz, q: 10
pole: f: 1.02Hz q: 10  / f: 3Hz, q: 5  / f: 2Hz, q: 6

FM9: (HF Roll-off)
pole: f: 40Hz q: 1/sqrt(2)
 
Servo gain: -0.027


The loop gains were adjusted to have the UGFs of 10Hz. The measured openloop transfer functions were compared with the model.
The transfer functions for yaw are well matched. However, the pitch ones don't. It seems that the pitch loop has extra low pass
which I can't locate. The possibility is the analog electronics of the pitch loop.


The effect of the control between 0.3Hz to 3Hz are well represented by the model. The attachment 2 shows the free running
angle fluctuation, the ones with the control engaged, and the estimated spectra. Indeed, the estimated spectra well represent
the measured angular spectra.

Attachment 1: PRM_QPD.pdf
PRM_QPD.pdf
Attachment 2: QPD_spe.pdf
QPD_spe.pdf
  9930   Thu May 8 14:37:02 2014 JenneUpdateASCPOP ASC QPD power

I was thinking about POP today, and wanted to know if there was something to be done to allow us to use the PRCL ASC for at least a little bit farther into arm power buildup.

Anyhow, I checked, and while PRMI is locked on sidebands (ETMs misaligned), POP DC is about 80 counts, and the power measured by the Ophir power meter is 24 microWatts. 

We were on the 3rd gain setting for the QPD's power amplifier.  I turned it down to the "2" option.  (When at 4, the front panel light indicates saturation).

It's not clear to me what the gain settings mean exactly.  I think that "1" means 4*10^3 V/A, and "6" means 4*10^6 V/A (On-Trak OT301 info site), but I don't know for sure how the gain changes for the settings 2-5.  Anyhow, I have changed the digital gain for the ASC to be -0.063 from -0.023 for both pitch and yaw.

  10114   Mon Jun 30 22:06:55 2014 JenneUpdateASCIFO (mostly) aligned - ASS working

[Koji, Jenne]

The Yarm ASS is now working (as is the Xarm ASS).  Both of the TT's pitch servos had a sign flip.  We don't know why.

To start, we lowered the matrix elements that push on the TTs by a factor of 3, to compensate for the new factor of 3 in the slider gains:  ezcastep C1:ASS-YARM_OUT_MTRX_5_5 /3 C1:ASS-YARM_OUT_MTRX_5_7 /3 C1:ASS-YARM_OUT_MTRX_6_6 /3 C1:ASS-YARM_OUT_MTRX_6_8 /3 C1:ASS-YARM_OUT_MTRX_7_5 /3 C1:ASS-YARM_OUT_MTRX_7_7 /3 C1:ASS-YARM_OUT_MTRX_8_6 /3 C1:ASS-YARM_OUT_MTRX_8_8 /3

We turned off all 4 tip tilt ASS servos (in the Yarm ASS servo screen), and turned them on one at a time.  By doing this, we discovered that the pitch servos for both TT1 and TT2 needed to have the opposite sign from what they used to have.  However, the yaw servos kept the original signs.  It really doesn't make sense to me why this should be, but this is the way the ASS servo works.  We left both Xarm and Yarm ASSs on for several minutes, and saw that they didn't push any mirrors out of alignment.

The ASS_DOTHER_ON burt snapshot has been resaved with the new values.

 

Also, earlier this evening, I aligned the Yarm green beam to the cavity, although the cavity was not optimally aligned, so this needs to be re-done.

 

On our to-do list should be to add the tip tilt slider values to the DAQ channels list.

  10124   Wed Jul 2 16:18:32 2014 JenneUpdateASCIFO aligned, PRMI + 2 arms achieved

After the meeting, I aligned the IFO to the IR, and then I aligned the Ygreen to the Yarm.  I then found the beatnotes and used ALS to hold the arms with CARM/DARM, locked the PRMI, and reduced the CARM offset until I had arm powers of about 3.  Given that this was at 3pm, and people were tromping all over inside the IFO room, I feel positive about tonight.  

So, IFO seems ready, carm_cm_up script was successful, and got me to arm powers of 1, and then I further reduced the offset by a bit to go a little higher. 

  10278   Sat Jul 26 14:45:33 2014 GabrieleMetaphysicsASCResponse of POP QPD

 Koji asked me to perform a simulation of the response of POP QPD DC signal to mirror motions, as a function of the CARM offset. Later than promised, here are the first round of results.

I simulated a double cavity, and the PRC is folded with parameters close to the 40m configuration. POP is extracted in transmission of PR2 (1ppm, forward beam). For the moment I just placed the QPD one meter from PR2, if needed we can adjust the Gouy phase. There are two QPDs in the simulation: one senses all the field coming out in POP, the other one is filtered to sense only the contribution from the carrier field. The difference can be used to compute what a POP_2F_QPD would sense. All mirrors are moved at 1 Hz and the QPD signals are simulated:

pop_qpd_all.png

This shows the signal on the POP QPD when all fields (carrier and 55 MHz sidebands) are sensed. This is what a real DC QPD will see. As expected at low offset ETM is dominant, while at large offset the PRC mirrors are dominant. It's interesting to note that for any mirror, there is one offset where the signal disappears.

pop_qpd_carrier.png

This is the contribution coming only from the carrier. This is what an ideal QPD with an optical low pass will sense. The contribution from the carrier increases with decreasing offset, as expected since there is more power.

pop_qpd_sb.png

Finally, this is what a 2F QPD will sense. The contribution is always dominated by the PRC mirrors, and the ETM is negligible.

The zeros in the real QPD signal is clearly coming from a  cancellation of the contributions from carrier and sidebands.

The code is attached.

Attachment 4: foldeddoublecavity.mist
classname FoldedDoubleCavity

# parameters
const Pin  1                # input power
const Lprc 6.752            # power recycling cavity length
const d_BS_PR3 0.401        # folding mirror distances
const d_PR2_PR3 2.081
const d_PRM_PR2 1.876
const c 299792458           # speed of light
const fmod 5*c/(4*Lprc)     # modulation frequency, matched to Lprc
... 51 more lines ...
Attachment 5: pop_qpd.m
% compile simulation class
clear classes
m = MIST('foldeddoublecavity.mist');

% create simulation object
s = FoldedDoubleCavity(8);

% set angulat motion
s.PRM.setMotionShape('pitch');
s.PR2.setMotionShape('pitch');
... 85 more lines ...
  10337   Wed Aug 6 10:45:56 2014 GabrieleUpdateASCPOP QPD signals

 In addition to the simulation described in my previous elog, I simulated the signal on a quadrant photodetector demodulated at 2F. The input laser beam is modulated at 11MHz up to the fifth order. There is no additional 55 MHz modulation.

The QPD demodulated at 2F shows good signals for PRC control for all CARM offsets, as expected from the previous simulation.

pop_qpd_2f.png

 

  10338   Wed Aug 6 12:44:52 2014 KojiUpdateASCPOP QPD signals

This is nice. Can we test this idea with POP22 + a razor blade?

Just to take transfer functions in PRMIsb between the PRM angle to POP QPD/POP22+razor blade
as well as the noise spectrum measurement are already useful.

We want to figure out the requirement for the 2f QPD.
(Transimpedance / Noise level / Beam size / etc)

Depending on the requirement we'll see if we need demodulation or just a power detector.

  10398   Fri Aug 15 01:31:44 2014 JenneUpdateASCPOP QPD to ETMs model wiring complete

I have added a few things to the ASS model, and the ASC sub-block, so that we can send POP QPD information down to the ETMs for CARM angular control after we've reduced the CARM offset and gotten some carrier buildup.  I did not remove our ability to actuate on PRM, so that we can still play with it in PRMIsb cases.

The input matrix has been expanded so that it can send signals to new CARM_YAW and CARM_PIT filter banks.  The corresponding filter banks have been created.  The output matrix was also expanded to take in the 2 new servo outputs, and so it can send signals to both ETMs, pitch and yaw.  I did not include any triggering logic for this new CARM situation, since I assume we'll just turn it on and off with our scripts.  (We haven't really been using the triggering capability of the PRM ASC either lately, although it's all still there).  I added the inputs and outputs of the CARM servos to the list of acquired channels.

The ASC sub-block:

ASC_newCARM.png

I also modified the top level of the ASS model.  This was just a simple addition of summing nodes for the ETMs, similar to what was already in place for the PRM, so that we can send both the ASS dither alignment signals and the ASC servo control signals to the optics.

The ASS top level:

ASSoverview_newCARM.png

I also quickly modified the ASC screen to expose all of the new options:

ASC_screen.png

The ASS model was compiled, and restarted.  As usual, this temporarily removes the biases on the input pointing tip tilts, but the pointing seems to have come back without any trouble.

  10454   Thu Sep 4 18:30:13 2014 GabrieleSummaryASCOptimal Gouy phase for POP QPD

 Jenne asked me to simulate the signals on POP QPD when moving different mirrors, as a function of the Gouy phase where the QPD is placed.

I used the opportunity to create a MIST simulation file of the entire 40m interferometer, essentially based on my aLIGO configuration file. I used the recycling cavity lengths obtained from our survey, and other parameters from the wiki page. The configuration file is attached (fortymeters.mist).

Coming back to the main simulation, here is the result, both for the "regular" POP QPD and for a 22MHz demodulated one. The Gouy phase is measured starting from PR2. Cavity mirrors are easily decoupled from PRM in the "regular" QPD. As already demonstrated in a previous simulation, ETMs signals are very small in the 22 MHz QPD. Moreover, it is possible to zero the contribution from ITMs by choosing the right Gouy phase, at the price of a reduction of the PRM signal by a factor of 3-4. Simulation files are attached.

pop_qpd_dc.png

 

Attachment 2: fortymeters.mist
###########################################################################
# Configuration file for full dual recycled 40m interferometer
classname FortyMeters
################################################################ Parameters

# General parameters
const Pin 1             # input power

# Mirror parameters
const T_ITM 0.01384     # ITM transmission [from https://wiki-40m.ligo.caltech.edu/Core_Optics]
... 143 more lines ...
Attachment 3: fortymeters_pop_qpd.mist
###########################################################################
# Configuration file for full dual recycled 40m interferometer
classname FortyMetersPOP_QPD
################################################################ Parameters

# General parameters
const Pin 1             # input power

# Mirror parameters
const T_ITM 0.01384     # ITM transmission [from https://wiki-40m.ligo.caltech.edu/Core_Optics]
... 148 more lines ...
Attachment 4: pop_qpd.m
% compile and create simulation class
clear classes
MIST('fortymeters_pop_qpd.mist');
s = FortyMetersPOP_QPD(4);

% set angular motion of ITMs, ETMs and PRM
s.ETMX.setMotionShape('pitch');
s.ETMY.setMotionShape('pitch');
s.ITMX.setMotionShape('pitch');
s.ITMY.setMotionShape('pitch');
... 47 more lines ...
  10503   Fri Sep 12 15:10:09 2014 JenneUpdateASCMICH ASS

During the Sim meeting today, I added parts to the ASS model so that we can also dither the BS and minimize the power at AS. 

ASS screen has been updated. 

Model changes required a new sender from LSC for ASDC, so both LSC and ASS were compiled, installed and restarted.  Also on LSC, I added AS110 I&Q to DQ channels, since we haven't been recording them in the past.

I have done a quick trial in MICH-only lock, and it seems to work.  Gain of 10 for both Pit and Yaw servos. 

  10662   Mon Nov 3 17:14:00 2014 ericqUpdateASCIMC to IFO angular motion

Something to note, as we have the IMC angular controls under consideration:

Jenne has the DRMI locked right now. I took a look at the coherence between the POP QPD and MC2 transmission QPDs. (Since she's using ASC, I also included those control signals. The coherences are about the same, unsurprisingly)

Based on the observed coherences, from about 1 to 6Hz, IMC motion is responsible for a fair amount of the DRMI angular motion. Also, PIT and YAW couple differently. 

2014-10-03-MC2T_to_POPQPD.pdf

  10663   Mon Nov 3 17:43:14 2014 KojiUpdateASCIMC to IFO angular motion

I wonder if this is the coherence caused by the beam itself, or caused by the same ground motion.
Jenne should be able to tell us...

  10680   Thu Nov 6 12:53:09 2014 diegoUpdateASCX arm restored

[Diego, Koji]

X arm has been restored, after modifying the two parameters mentioned in http://nodus.ligo.caltech.edu:8080/40m/10676 (C1SUS_ITMX:  LSC/DAMP and LSC/BIAS); after that, a manual re-alignment of ETMX was necessary due to heavy PIT misalignment. I will check the ALS lock once work on the Y arm is done.

  10745   Tue Dec 2 01:27:22 2014 diegoUpdateASCASS Scripts for arms

I updated the medm C1ASS page for the Arm scripts:

ON : same as before

FREEZE OUTPUTS: calls new FREEZE_DITHER.py script, which sets Common Gain and LO Amplitudes to 0, therefore freezing the current output values

START FROM FROZEN  OUTPUTS: calls new UNFREEZE_DITHER.py script, which sets Commong Gain and LO Amplitudes as in the DITHER_ASS_ON.py script, but no burt restore is performed

OFFLOAD OFFSETS: it's the old "SAVE OFFSETS", calls the WRITE_ASS_OFFSET.py script

OFF: same as before

StripTool: same as before

 

 

  10752   Thu Dec 4 00:26:07 2014 JenneUpdateASCPOP yaw razor blade installed

We would like the option of feeding back the POP beam position fluctuations to the PRM to help stabilize the PRC since we don't have oplevs for PR2 and PR3.  However, we cannot just use the DC QPD because that beam spot will be dominated by carrier light as we start to get power recycling. 

The solution that we are trying as of today is to look at yaw information of just the RF sidebands.  (Yaw is worse than pitch, although it would be nice to also control pitch).  I have placed a razor blade occluding about half of the POP beam in front of the POP PD (which serves POPDC, POP22 and POP110).  I also changed the ASS model so that I could use this signal to feed back to the PRM.  Loop has been measured, and in-loop spectra shows some improvement versus uncontrolled.


Optical table work:

The POP beam comes out of the vacuum system and is steered around a little bit, then about 50% goes to the DC QPD.  Of the remaining, some goes to the Thorlabs PD (10CF I think) and the rest goes to the POP camera.  For the bit that goes to the Thorlabs PD, there is a lens to get the beam to fit on the tiny diode.

There was very little space between the steering mirror that picks off the light for this PD, and the lens - not enough to put the razor blade in.  The beam after the lens is so small that it's much easier to occlude only half of the beam in the area before the lens.  (Since we don't know what gouy phase we're at, so we don't know where the ideal spot for the razor is, I claim that this is a reasonable place to start.)

I swapped out the old 50mm lens and put in a 35mm lens a little closer to the PD, which gave me just enough room to squeeze in the razor blade.  This change meant that I had to realign the beam onto the PD, and also that the demod phase angles for POP22 and POP110 needed to be checked.  To align the beam, before placing the razor blade, I got the beam close enough that I was seeing flashes in POPDC large enough to use for a PRMI carrier trigger.  The PRMI carrier was a little annoying to lock.  After some effort, I could only get it to hold for several seconds at a time.  Rather than going down a deep hole, I just used that to roughly set the POP22 demod phase (I -phase maximally negative when locked on carrier, Q-phase close to zero).  Then I was able to lock the PRMI sideband by drastically reducing the trigger threshold levels.  With the nice stable sideband-locked PRMI I was able to center the beam on the PD. 

After that, I introduced the razor blade until both POPDC and POP22 power levels decreased by about half. 

Now, the POP22 threshold levels are set to up=10, down=1 for both MICH and PRCL, DoF triggers and FM triggers.


ASS model work:

POP22 I and POP110 I were already going to the ASS model (where ASC lives) for the PRCL ASS dither readbacks.  So, I just had to include them in the ASC block, and increased the size of the ASC input matrix.  Now you can select either POP QPD pit, POP QPD yaw, POP221 or POP110I to go to either PRCL yaw, PRCL pit, CARM yaw or CARM pit. 

Compiled, installed and restarted the ASS model.


Engaging the servo:

I took reference spectra of POP QPD yaw and POP 22, before any control was applied.  The shapes looked quite similar, but the overall level of POP22 was smaller by a factor of ~200.  I also took a reference spectra of the POP QPD in-loop signal using the old ASC loop situation.

Q looked at Foton for me, and said that with the boost on, the UGF needed to be around 9 or 10 Hz, which ended up meaning a servo gain of +2.5 (the old POP QPD yaw gain was -0.063).  We determined that we didn't know why there was a high-Q 50Hz notch in the servo, and why there is not a high frequency rolloff, so right now the servo only uses FM1 (0:2000), FM6 (boost at 1Hz and 3Hz) and FM7 (BLP40). 

The in-loop residual isn't quite as good with POP22 as for the QPD, but it's not bad. 

Here's the loop:

ASC_PRCLloop_POP22err.pdf

And here's the error spectra.  Pink solid and light blue solid are the reference traces without control.  Pink dashed is the QPD in-loop.  Red and blue solid are the QPD and POP22 when POP22 is used as the error signal.  You can definitely see that the boosts in FM6 have a region of low gain around 1.5Hz.  I'm not so sure why that wasn't a problem with the QPD, but we should consider making it a total 1-3Hz bandpass rather than a series of low-Q bumps.  Also, even though the POP22 UGF was set to 9 Hz, we're not seeing any suppression above about 4Hz, and in fact we're injecting a bit of noise between 4-20Hz, which needs to be fixed still. 

PRC_YAW_QPDvs22_3Dec2014.pdf

  10768   Tue Dec 9 03:34:52 2014 JenneUpdateASCPOP yaw razor tuning

With the re-do of the IFO alignment last week, I think that the beam was no longer about halfway on the POP22 razor blade.  To fix this, I locked the PRMI on sideband, removed the razor blade, and then put it back in such that it occluded about half of the light.  

I'm not entirely sure why, but when I put the razor in, POP22 went from 104(ish) to 45(ish) but POPDC  went from 5200(ish) to 1600(ish).  [The 'ish'es are because the PRC wasn't angularly stabilized, so there was some motion changing the power levels that leaked out to the POP port].  The ETMs were misaligned, so this should not be a carrier vs. sideband effect, since they'll both share the cavity axis defined by the ITMs and the PRM.  It is possible, although I didn't check, that there is some oplev light scattered into the POP photodiode that is now blocked by the razor blade.  This light would only be at DC and not the 2f frequencies.  Since the signal levels for POP22 vs. POPDC didn't change with and without the table top on (and with and without room lights on), I don't think that it is an effect of ambient light getting into the diode.  To check if it is oplev light I should (a) just look, and (b) try to lock the PRMI without the ITMX oplev laser being on to see if there is a difference in the POPDC signal.

Anyhow, under the assumption that the POP22 signal level is correct, I tuned up the PRCL ASC a little bit.  These changes are now in the carm_cm_up script, and the carm_cm_down script resets things.  Before the PRC is locked, I have FM1 and FM7 (the basic servo shape and a 40Hz lowpass) on, the gain set to zero, and the input off.  After lock is acquired, the input is turned on, and the gain ramps from 0 -> 10 in 3 seconds.  Then FM2 and FM6 (boosts at 1 and 3Hz) are engaged.

In the plot below, the dark blue and red curves were taken when there was no angular control on the PRC.  Pink was taken last week with the old QPD yaw ASC on.  Light blue is today's version of the in-loop performance of the POP22 yaw ASC loop.  I didn't save the trace unfortunately, but the DC QPD saw out-of-loop improvement between about 0.8Hz - 4 Hz. 

Also, has anything happened with the LSC rack in the last few weeks that might be causing lots of 60Hz noise? I saw these large lines last week, but I don't think I remember them from the past.

PRC_YAW_QPDvs22_8Dec2014.pdf

After I got the PRCL ASC working, I tried several iterations of locking.  ETMX is still being annoying, although the last hour or so have been okay.  CARM keeps getting rung up right around the transition to the sqrtInv error signal.  Since CARM and DARM are kind of entangled, it took me a few iterations to figure out that it was CARM that is ringing up, and not DARM.  I'm a little worried about the phase loss from the 1kHz lowpass that we turn on just before the transition to sqrtInv.  I want to keep the lowpass off until after we have transitioned DARM also over to DC transmission.  I tried once, but I lost lock before starting the CARM transition.  Anyhow, the ETM alignment issue is annoying.

Also, Jamie, Q, Diego and I were discussing last Friday, but none of us elogged, that we think there might be something wrong with one of the Martian network switches.  I'll start a separate thread about that right now, but it slows things down when you can't trust EPICS channels to be current, and I (without evidence) am a little worried that this might also affect the fast signals.

  10789   Fri Dec 12 04:33:49 2014 JenneUpdateASCASS retuned

[Rana, Jenne]

We decided that tonight was the night for ASS tuning. 

We started from choosing new frequencies, by looking at the transmission and the servo control signals spectra to find areas that weren't too full of peaks.  We chose to be above the OpLev UGF by at least a factor of ~2, so our lowest frequency is about 18Hz.  This way, even if the oplevs are retuned, or the gains are increased, the ASS should still function. 

We set the peak heights for the lowest frequency of each arm to have good SNR, and then calculated what the amplitude of the higher frequencies ought to be, such that the mirrors are moving about the same amount in all directions. 

We re-did the low pass filters, and eliminated the band pass filters in the demodulation part of the servo.  The band passes aren't strictly necessary, as long as you have adequate lowpassing, so we have turned them off, which gives us the freedom to change excitation frequencies at will.  We modified the lowpass filter so that we had more attenuation at 2Hz, since we spaced our excitation frequencies at least ~2.5 Hz apart.

The same lowpass filter is in every single demodulator filter bank (I's and Q's, for both length and transmission demodulation).  We are getting the gain hierarchy just by setting the servo gains appropriately. 

We ran ezcaservos to set the demodulation phase of each lockin, to minimize the Q-phase signal. 

We then tuned up the gains of the servos.  Rana did the Y arm, but for the X arm I tried to find the gains where the servos went unstable, and then reduced the gain by a factor of 2.  The Xarm is having trouble getting good alignment if you start with something less than about 0.7, so there is room for improvement.

Rana wrote a little shell script that will save the burt snapshot, if the gains need adjusting and they should be re-saved. 

The scripts have been modified (just with the new oscillator amplitudes - everything else is in the burt snapshots), so you should be able to run the start from nothing and the start from frozen scripts for both arms.  However, please watch them just in case, to make sure they don't run away.

  10807   Wed Dec 17 01:51:44 2014 rana, jenneUpdateASCASS retuned

Did a big reconfig to make the Y-arm work again since it was bad again.

  1. Undid Koji's topology change. The A2L loops now feedback to the arm mirrors to adjust the cavity axis. The cavity transmission signals now feedback to the input beam.
  2. The UGF of the Trans->Input beam servos is ~5-10x higher than the A2L servos.
  3. The Trans loops have a ~10-15 s settling time.
  4. The Input Matrix has been adjusted to fit with our intuition:The ETM tilt moves the beam equally on the ITM and ETM faces.
  5. The Output Matrix has also been adjusted to do like this: we're using an intuitive matrix inverse rather than one based on measurement. It turns out to be a reasonable guess and we can tune this later.
  6. Seems stable with many kinds of steps and misalignments. Seems not reliable if the arm power is less than ~0.5.
  7. Reducing the dither amplitudes to make the power fluctuation less than 5% made it much more stable.

With the arm aligned and the A2L signals all zeroed, we centered the beam on QPDY (after freezing the ASS outputs). I saw the beam going to the QPD on an IR card, along with a host of green spots. Seems bad to have green beams hitting the QPD alogn with the IR, so we are asking Steve to buy a bunch of the broad, dielectric, bandpass filters from Thorlabs (FL1064-10), so that we can also be immune to the EXIT sign. I wonder if its legal to make a baffle to block it on the bottom side?

P.S. Why is the Transmon QPD software different from the OL stuff? We should take the Kissel OL package and put it in place of our old OL junk as well as the Transmons.

Attachment 1: ASSconfig_141217_0205.png
ASSconfig_141217_0205.png
  10811   Wed Dec 17 18:14:36 2014 ericqUpdateASCTransmon QPD -> ASC servos ready for comissioning

 I have completed all of the model modifications and medm screen updates to allow for feedback from the transmon QPD pitch and yaw signals to the ITMs. Now, we can design and test actual loops...

newASCscreen.png

The signals come from c1sc[x/y] to c1rfm via RFM, and then go to c1ass via dolphin. 


Out of curiosity about the RFM+dolphin delay, I took a TF of an excitation at the end SUS model (C1:SUS-ETM[X/Y]_QPD_[PIT/YAW]_EXC) to the input FM in the ASC model (C1:ASC-ETM[X/Y]_QPD_[PIT/YAW]_IN1). All four signals exhibit the same delay of 122usec. I saved the dtt file in Templates/ASC/transmonQPDdelay.xml

This is less than a degree under 20Hz, so we don't have to worry about it. 

  10812   Wed Dec 17 19:04:12 2014 jenneUpdateASCASS retuned

I made the Xarm follow the new (old) topology of Length -> test masses, and Trans -> input pointing.

It takes a really long time to converge (2+ min), since the input pointing loops actuate on the BS, which has an optical lever, which is slow.  So, everything has to be super duper slow for the input pointing to be fast relative to the test mass motion.

Also, between last night and this afternoon, I moved the green ASX stuff from a long list of ezca commands to a burt file, so turning it on is much faster now.  Also, I chose new frequencies to avoid intermodulation issues, set the lockin demodulation phases, and tuned all 4 loops.  So, now the green ASX should work for all 4 mirrors, no hand tuning required.  While I was working on it, I also removed the band pass filters, and made the low pass filters the same as we are using for the IR ASS.  The servos converge in about 30 seconds.

  10813   Wed Dec 17 19:31:55 2014 KojiUpdateASCASS retuned

I wonder what to do with the X arm.

The primary purpose of the ASS is to align the arm (=transmission), and the secondary purpose is to adjust the input pointing.

As the BS is the only steering actuator, we can't adjust two dof out of 8 dof.
In the old (my) topology, the spot position on ITMX was left unadjusted.

If my understanding of the latest configuration, the alignment of the cavity (=matching of the input axis with the cavity axis)
is deteriorated in order to move the cavity axis at the center of the two test masses. This is not what we want as this causes
deterioration of the power recycling gain.

  10886   Mon Jan 12 18:11:25 2015 ericqUpdateASCTest Mass -> Transmon QPD TFs measured

We want to have some angular control of the arms during lock acquistion. 

In single arm lock, Diego and I shook the TMs and measured how the QPDs responded. (I would've liked to do a swept sine in DTT, but the user envelope function still isnt' working!)

For now, we can close simple loops with QPD sensor and ITM actuator, but, as Rana pointed out to Diego and me today, this will drive some amount of the angular cavity degree of freedom that the QPD doesn't sense. So, ideally, we want to come up with the right combination of ITM and ETM motion that lies entirely within the DoF that the QPD senses.

I created a rudimentary loop for Yarm yaw, was able to get ~20Hz for the upper UGF, a few mHz for the lower, but it was starting to leak into the length error signal. Further tweaking will be neccesary...

Attachment 1: Jan12_singleArmSensing.pdf
Jan12_singleArmSensing.pdf
Attachment 2: Jan12_singleArmSensing.xml.zip
  10908   Thu Jan 15 18:57:41 2015 ericqUpdateASCTransmon QPD loops live

I've measured the sensing for each of the arms, by using our calibrated oplevs, in terms of QPD counts per micron. It is:

YARM
ETMY: QPD PIT / OPLEV PIT =   22.0 count/urad
      QPD YAW / OPLEV YAW =   17.1 count/urad
ITMY: QPD PIT / OPLEV PIT =   -6.0 count/urad
      QPD YAW / OPLEV YAW =    5.9 count/urad
XARM
ETMX: QPD PIT / OPLEV PIT =   16.6 count/urad
      QPD YAW / OPLEV YAW =   -9.3 count/urad
ITMX: QPD PIT / OPLEV PIT =    4.0 count/urad
      QPD YAW / OPLEV YAW =   -6.0 count/urad

In the absence of a lens, the QPD would be significantly more sensitive to cavity axis translation than tilt, and thus about equally sensitive to ITM and ETM angle. However, there are lenses on the end tables. I didn't go out and look at them, but found some elogs from Annalisa that mentioned 1m focal length lenses. Back-of-the-envelope calculations convince me that this can plausibly lead to the above sensitivity ratios.

I used these quantities to come up with an actuation matrix for the ASC loops, and measured the effective plant seen by the FM, fitted it to some poles( looks like zpk([],-2*pi*[1.47+3.67i,1.47-3.67i],160); ), and designed a control servo. Here is the designed loop:

The servo works on both arms, both DoFs. A DTT measurement agrees with the designed loop shape, up to a few degrees, which are probably due to the CDS delay. The RMS of the QPD error signals goes down by about 20dB, and are currently dominated by the bounce mode, so maybe we can try to sneak in some resonant gain...?

Once we confirm that they work when locking, we can write up and down lines into the locking scripts...

Attachment 1: loopDesign.pdf
loopDesign.pdf
  10920   Mon Jan 19 18:27:16 2015 ericqUpdateASCQPD ASC saga continues.

Herein, I will try to paint a more thorough picture of this whole QPD ASC mess. 


Motivation for QPD ASC loops:

  • We would like to use the QPDs as a DC arm pointing reference during locking attempts, or over multiple days, if possible. 
  • It would be nice if the QPDs could complement the oplevs to reduce angular motion of the cavity. 
  • We must not make the single arm longnitudinal noise or RIN worse, because anything observable in the single arm case will be catastrophic at full sensitivity

Actuation design:

As mentioned in a previous ELOG, in single arm lock, I measured the QPD response with respect to the calibrated oplev signals. They were:

YARM
ETMY: QPD PIT / OPLEV PIT =   22.0 count/urad
      QPD YAW / OPLEV YAW =   17.1 count/urad
ITMY: QPD PIT / OPLEV PIT =   -6.0 count/urad
      QPD YAW / OPLEV YAW =    5.9 count/urad
XARM
ETMX: QPD PIT / OPLEV PIT = 16.6 count/urad
      QPD YAW / OPLEV YAW = -9.3 count/urad
ITMX: QPD PIT / OPLEV PIT =  4.0 count/urad
      QPD YAW / OPLEV YAW = -6.0 count/urad

For reference, one microradian of either ITM or ETM motion produces about 60um of ETM beam spot displacement, compared to the spot size of ~5mm. 

However, given the lenses on the end tables that are used for green mode matching, that the IR transmitted beam also passes through, the QPDs are not directly imaging the ETM spot position; if they were, they would have equal sensitivity to ITM and ETM motion due to our flat/curved arm geometry. 

From this data, I calculated the actuation coefficients for each DoF as A_{ETM} = \frac{d_{ETM}}{\sqrt {d_{ETM}^2 + d_{ITM}^2}}, and similarly for the ITMs, where the d's come from the table above. However, it occurs to me that maybe this isn't the way to go... I'll mention this later. 


Loop design:

Up until now, at every turn, I had not properly been thinking about how the oplev loop plays into all of this. I went to the foton filters, and grabbed the loop and plant models for the ETMY oplev, and constructed the closed loop gain, 1/1+G, and the modified plant, P/1+G, which is what the ASC loop sees as its plant. 

Here, the purple trace explains all of the features I was confused about earlier. 

With this in hand, I set up to design a loop to satisfy our motivations. 

  • Bounce/roll mode notches to avoid exciting them
  • 1/f UGF crossing at a few Hz, limited by the gain margin at ~10Hz, which is where the phase will hit 180, due to the notches
  • 4th order Elliptic lowpass at 100, to avoid contaminating the longnitudinal signals
  • 1/f^2 at low frequencies for DC gain

To do this, I inverted the oplev closed loop plant pole around 4Hz to smooth the whole thing out. Here's a comparison of the measured OLG with what I modelled. 

There's a little bit of phase discrepency around 10Hz, but I think it looks about right overall. 


Evaluation:

So, here's the part that counts: How does this actually perform? I took spectra of the QPD error signals, the relevant OpLev signals as out of loop sensors, the PDH error signal and transmitted RIN while single arm locked, with this loop off, and on for all 4 DoFs simultaneously. 

Verdict:

  • In-loop signals get small, unsurprisingly.
  • Cavity signals unchanged. yes
  • ITM oplev signals are unchanged (and not plotted, to not clutter the plots (This isn't surprising since the loops mostly actuate on the ETMs).
  • ETM oplev signals get smaller around the 3Hz peak, but are louder in other bands.no 

This is what makes me think I may need to revisit the actuation matrix. If I did it wrong, I am driving the "invisible" quadrant of the cavity angular DoFs, and this could be what is injecting noise into the oplevs. 


Conclusion:

In the end, I have a better understanding of what is going on, and I don't think we're quite there yet.  

Attachment 1: oplevPlant.pdf
oplevPlant.pdf
Attachment 2: loopDesignComparison.pdf
loopDesignComparison.pdf
  10921   Tue Jan 20 02:39:49 2015 ericqUpdateASCQPD ASC saga continues.

Although the QPD loops are less than ideal right now, I made changes to the ASC model to trigger the QPD loops on and off politely, depending on TRX and TRY. The settings are exposed on the ASC screen. However, I have not yet exposed the FM triggering that I also set up to make sure the integrator doesn't misbehave if the arm loses lock. We probably don't want to trigger them on at anything lower than arm powers of about 1.0. 

I've tested the triggering by randomly turning LSC mode on and off, and making sure that the optics don't recieve much of a kick as the QPD loops engage a few seconds after the LSC boosts do, or when lock is lost. This works as long as there isn't much of a DC offset befire the loops are engaged. (Under 20 counts or so is fine)

As a side note, I was going to use the TRIG_SIG signals sent via the LSC model via SHMEM blocks for the ASC triggering, but oddly, the data streams that made it over were actually the MICH and SRCL TRIG_SIGs, instead of XARM and YARM as labelled. I double checked the simulink diagrams; everything seemed fine to me. In any case, ASS was already recieving TRX and TRY directly via RFM, so I just piped those over to the ASC block. This way is probably better anyways, because it directly references the arm powers, instead of the less obvious LSC triggering matrix. 

  10946   Tue Jan 27 21:33:39 2015 KojiUpdateASCASS retuned

I checked the situation of ASS. I wanted to know how much we are away from the maximum transmittion.

Conclusion:
ASS makes the X arm shifted from the maximum transmission. This causes the contrast degraded by ~3%.
We need to fix the Xarm ASS so that it can maximize the transmission and ignor the spot centering at ITMX.


Conditioning before the measurement:

- ASDC offset was removed
- X&Y arm was aligned by ASS

With ASS:

Average transmission: 0.86
Pmax = 1045 +/- 9 cnts
Pmin = 22 +/- 4 cnts

==> Contrast = (Pmax - Pmin)/(Pmax+Pmin) = 0.960+/-0.007

After manual alignment of the X arm (ignoring spot centering):

Average transmission: 0.88
Pmax = 1103 +/- 11 cnts
Pmin = 5 +/- 1 cnts

==> Contrast = (Pmax - Pmin)/(Pmax+Pmin) = 0.991+/-0.002

  10952   Wed Jan 28 23:53:24 2015 KojiSummaryASCXarm ASS fix

X-Arm ASS was fixed.
ASS_DITHER_ON.snap was updated so that the new setting can be loaded from the ASS screen.

The input and output matrices and the servo gains were adjusted as found in the attached image.
The output matrix was adjusted by looking at the static response of the error signals when a DC offset
was applied to each actuator.

The servo was tested with misalignment of the ITM, ETM, and BS. In fact, the servo restored transmission
from 0.15 to 1.

The resulting contrast after ASSing was ~99% level. (I forgot to record the measurement but the dark fringe level of ASDC was 4~5count.)

Attachment 1: 12.png
12.png
  10963   Mon Feb 2 12:24:27 2015 JenneUpdateASCPOP QPD centered

After aligning the PRC, I centered the POP QPD.

  10974   Wed Feb 4 18:27:55 2015 KojiSummaryASCXarm ASS fix

Please remember that Xarm ASS needs FM6 (Bounce filters) to be ON in order to work properly.

  10975   Wed Feb 4 19:21:37 2015 KojiUpdateASCArm ASS servos now have triggered gain with arm lock status

We had persistent frustration by occasional unlock during ASSing.
Today, I added triggers to the servo gains in order to elliminate this annoyance.

Each ASS servo gain slider is multiplied with the corresponding LSC Trigger EPICS channel (i.e. C1:LSC-iARM_TRIG_MON, where i=X or Y).
This has been done by ezcaread modules in RCG.

The model and screen have been commited to svn.

  10997   Tue Feb 10 18:37:17 2015 ericqUpdateASCQPD ASC ready to go

I've remeasured the QPD ASC sensing coefficients, and figured out what I did weird with the actuation coefficients. I've rearranged the controller filters to be able to turn on the boost in a triggered way, and written Up/Down scripts that I've tested numerous times, and Jenne has used as well; they are exposed on the ASC screen. 

All four loops (2 arms * pit,yaw), have their gains set for 8Hz UGF, and have 60 degrees of phase margin. The loop shape is the same as the previous ELOG post. Here is the current on/off performance. The PDH signals (not shown, but in attached xml) show no extra noise, and the low frequency RIN goes down a bit, whic is good. The oplevs error signals are a bit noisier, but I suppose that's unavoidable. The Y-arm performs a bit better than the X-arm. 

The up/down scripts don't touch the filters' trigger settings at all, just handles switching the input and output and clearing history. FM1 contains the boost which is intended to have a longer trigger delay than the filters themselves.

Attachment 1: Feb10_loops_offOn.png
Feb10_loops_offOn.png
Attachment 2: Feb10_newLoops_offOn.xml.zip
  11056   Fri Feb 20 19:09:48 2015 ericqUpdateASCQPD frontend code unified

I have changed all of the oplevs and transmon QPDs to use the common ISC QPD library block, which differs mainly in its divide by zero protection. 

c1scx.mdl and c1scy.mdl were directly changed for the transmon QPDs. The oplevs were done by changing the sus_single_control.mdl library part, which is used for all of the SOSs. 

Then, because of the underscore introduced (i.e. OLPIT becomes OL_PIT because there is an OL block), I went on a sed safari to find and replace the new channel names into:

  • The filter ini files
  • various MEDM screens
  • The optic misaligning scripts (which currently live in medm/MISC/ifoalign, and need to get moved to scripts/)
  • A recent BURT snapshot, to restore all of the switches and settings easily. 
  • scripts/activateDQ.py, which is responsible for renaming OL_PIT_IN1_DQ to OPLEV_PERROR, etc.

I've fixed everything that occured to me, and the usual ways I'm used to interacting with the oplevs all seem to work at this time, but it's entirely possible I've overlooked something.

One important note is: because we are now using an effectively immutable QPD library block, the oplev urad conversion has to take place in the DoF matrix. The EPICS records C1:SUS-[OPTIC]_OL_[DOF]_CALIB still exist, but do not multiply the fast signals. Rather, the OL_MTRX elements are multiples of the CALIB value. I thought about making a new QPD_CALIBRATED part or something, but then we're right back to using custom code, which is what we're trying to avoid. 

All of the oplev DoFs are stable, I checked a few loop TFs like ETMY pitch and PRM yaw, and they looked normal. 

  11057   Mon Feb 23 08:06:59 2015 ranaUpdateASCQPD frontend code unified

Might have to also get the OL screens that go with this new code to see, but the calibrations don't go into the matrix, but rather into the OLPIT/OLYAW filter banks which follow the division, but before the servo filter banks.

  11061   Tue Feb 24 18:54:26 2015 ericqUpdateASCSingle arm QPD ASC stability

I've lowered the UGFs for the transmission QPD servos to ~1-2Hz, and made it just an integrator. I left the arms locked with the QPD servos on for a few hours during the daytime today, and they succesfully prevented the Y arm from losing power from alignment drift for ~4 hours. Turning the servo off caused TRY to drop to ~0.6 or so. 

The X arm was only held for 2 hours or so, because after some unlock/drift event the power was below the servo trigger threshold. However, after gently nudging ETMX to get the transmission above the threshold, the servo kicked in, and brought it right back to TRX=1.0

Unfortunately, daqd was dead for much of the day, so I don't have much data to show; the trend was inferred from the wall striptool. 

It is not proven that there aren't further issues that prevent this from working with higher / more dynamic arm powers, but this is at least a point in favor of it working. 

EDIT: Here's a screenshot of the wall StripTool. Brown is TRY, blue is TRX. The downturn at the very end is me deactivating the servos. 

There is no scientific justifcation for the 0.9 threshold. Really, I should look at the noise/SNR again, now that there is some ND filtering on the QPDs. 

Attachment 1: trend.png
trend.png
  11125   Mon Mar 9 18:10:59 2015 JenneUpdateASCOplev filters re-copied

Back on Feb 20th (elog 11056) Q replaced all of our oplev parts with the aLIGO version. 

Unfortunately, after this it has seemed like there was something not quite right with the optical lever servos.

  • When we would restore kicked optics, after the osem rms values come down the scripts try to engage the optical levers.  This would often kick and ring up the optics (I've seen this with ETMX and ETMY, and once or twice with the beam splitter)  Not good.
  • Sometimes, if the optic wasn't kicked, it would oscillate.  I would see this in particular with ETMY, by seeing the green transmission at the PSL table oscillating.
  • Q noticed that sometimes when the oplevs' outputs were turned off, the outputs were railed at the limiter values.

Since, when the models were changed which gave us an extra underscore in the oplev names, Q did a find-and-replace in the foton text files, I was worried that this might have broken things.  I'm not entirely sure how it would have broken them (I didn't see any difference in a diff), but I've heard enough horror stories about the delicacy of the foton text files.

Anyhow, I opened the last archived foton files from just before Q made the change, and copy-and-pasted the design strings from the old filter banks to the new ones.  Hopefully this fixes things.

  11206   Tue Apr 7 04:21:45 2015 ericqUpdateASCAngular Control during Locking

[J, Q]

Alignment is making it tough for locks to last more than 10 minutes. Many (but not all) locklosses correlate with some optic drifting away, and taking all of the light with it. The other locklosses are the quick ones that seem to pop up out of nowhere; we haven't made any headway on these. We wanted to get to a state where we could just let the interferometer sit for some minutes, to explore the data, but got caught up with alignment and PRMI things.

We're finding that both ITMs experience some DC force when entering full PRFPMI lock. I will calculate the torque expected from radiation pressure + offset beam spot, especially for ITMX, where we choose the spot position to be uncontrolled by ASS. 

I set up the QPD ASC servos to act in a common/differential way on the ETMs. The C1:ASC-XARM_[PIT/YAW] filter modules act on the common alignment, whereas the C1:ASC-YARM_[PIT/YAW] filter modules act on the differential alignment. This can soon be cleaned up with some model renaming to reduce confusion. 

Using DC oplev values as a guide, we are hand tuning ITM alignment once the AO path is engaged and we see the DC drift occurring. Then, we set the QPD servo offsets and engage them. 

In this manner, we were able to lock the interferometer at:

  • Arm transmission 150 x single arm power
  • POPDC indicated a recycling gain of ~5.5
  • ASDC/POPDC indicated a contrast of 99.8%
  • REFLDC indicated a visibility of 80%

We made the PRMI transition to 1f numerous times, but found that the sideband power fluctuations would get significantly worse after the transition. 

We found that the gains that were previously used were too small by a factor of a few. There is a DC change visible in REFL165 before and after the transition (Also POP55, aka REFL55, is not DQ'd angry). Really, it isn't certain that we've zero'd the offset in the CARM board either, so REFL55's zero crossing isn't necessarily more trustworthy that REFL165's. We can go back in the data and do some 2D histograming to see where in the error signal space the sideband power is maximized. 

Jenne reports:

  • The all RF transition succeeded 13/29 times. 
  • PRMI 1f transision succeeded 10/10 times. 