  8015   Wed Feb 6 19:59:35 2013 ranaUpdateLockingPRC cavity gains

  Getting closer, but need to use the real measured AR reflectivity values, not the 1500 ppm guess. These should be measured at the correct angles and pol, using an NPRO.

  8029   Fri Feb 8 00:23:33 2013 ranaSummaryGeneralPRC/arm mode matching calculations

 

 I would guess that either flipping PR2 or PR3 would give nearly the same effect (g = 0.9) and that flipping both makes it even more stable (smaller g). But what we really need is to see the cavity scan / HOM resonance plot to compare the cases.

The difference of 0.5% in mode-matching is not a strong motivation to make a choice, but sensitivity to accidental HOM resonance of either the carrier or f1 or f2 sidebands would be. Should also check for 2*f2 and 2*f1 resonances since our modulation depth may be as high as 0.3. Accidental 2f resonance may disturb the 3f error signals.

  8051   Sat Feb 9 19:34:34 2013 ranaUpdateOpticsG&H - AR Reflectivity

 

 Use the trick I suggested:

Focus the beam so that the beam size at the detector is smaller than the beam separation. Use math to calculate the beam size and choose the lens size and position. You should be able to achieve a waist size of < 0.1 mm for the reflected beam.

  8096   Sun Feb 17 19:27:19 2013 ranaUpdateSUSPRM coil balanced

 I will check out the AS55 situation tomorrow. Just put it on my desk.

MC Autolocker was disabled - I enabled it.

For the F2P.py, you should look at how we did this with the script written 8 years ago in csh. There we stored the initial values in a file (so they don't get blow away if someone does CTRL-C). Your python script should have a trap for SIGINT so that it dies gracefully by restoring the initial values. In order to have the smooth value adjustment, you must first set the TRAMP field for all the coil gains to 2 and then switch. Make sure that the lockin ignores the first few seconds of data after making this switch or else it will be hugely biased by this transient.

For the PRM OL use as a F2A reference, you also have to take into account that the OL beam is hitting the PRM surface at non-normal incidence. IF it is a large angle, there will be a systematic error in the setting of the F2Y values.

  8120   Wed Feb 20 19:58:59 2013 ranaUpdateAlignmentBS table oplev re-arranged

Please confirm the SRM OL beam is not too bad and also find where the mis-aligned SRM puts its beam. WE want to be sure that there is not too much unwanted scattering from SRM into the PRFPMI.

  8123   Wed Feb 20 21:12:37 2013 ranaUpdateAlignmentclipping centering checklist

 

 Is the beam going towards the OMC going to cause backscatter because of uncontrolled OMC or can we park that beam somewhere dark?

  8132   Thu Feb 21 18:10:13 2013 ranaUpdateSEIPump Down misalignments

This plot shows the trend of the OL during the past several hours of roughing pumping.

The big steps at the start of the pump down is NOT due to the pumping, but is instead the "recentering" that Yuta did. Looks like he was unable to find zero on the ETMY.

Some of the rest of the drift is probably just the usual diurnal variation, but there does seem to be some relation to the pumping trend. I guess that the shift of ~0.3 in the ITMX and ITMY pitch is real and pressure related.

We need to figure out how to put the OL calibration factor into the SUS-OL screens.

 

Attachment 1: OL.png
OL.png
  8147   Sat Feb 23 15:46:16 2013 ranaUpdateComputerscrontab in op340m updated

According to Google, you can add a line in the crontab to backup the crontab by having the cronback.py script be in the scripts/ directory. It needs to save multiple copies, or else when someone makes the file size zero it will just write a zero size file onto the old backup.

  8153   Sun Feb 24 16:49:00 2013 ranaSummaryElectronicsReplacement for the AD743: OPA140 and OPA827

  This looks pretty good already. Not sure if we can even measure anything reasonable below 0.1 Hz without a lot of thermal shielding.

The 10-20 kHz oscillation may just be the loop shape of the opamp. I think you saw similar effects when using the AD743 with high impedance for the OSEM testing.

  8154   Sun Feb 24 17:54:34 2013 ranaUpdateSUSSUS Summary

 

 I asked John Z to talk with Jamie and then install a new NDS2 server software for us. Jamie may know if this happened or was foiled by the linux1 RAID failure.

In any case, our pyNDS stuff ought to be able to talk to NDS2 or our old NDS1 stuff, I hope.

  8158   Mon Feb 25 17:58:28 2013 ranaUpdateAlignmentY arm locked, both colors

 

 That's good news. I was ready to give up and say we should vent and remove the baffles. It will be interesting if you can find out how much the sensors and OL and IPANG are off from their pre-pump values. We should think about how to have better references.

Also, what is the story with the large drift we are seeing in IPANG?

  8191   Wed Feb 27 20:10:43 2013 ranaUpdateLockingDRFPMI flashes

 

 If its true that there have been large flashes, then there indeed might be beer. But first I'd have to see a calibrated plot. And make sure that the flashes are not overamplified due to a whitening filter imbalance.

Is it the readout of a PD with no whitening/antiwhitening? IF so, its much easier to believe.

  8195   Wed Feb 27 23:19:54 2013 ranaUpdateSummary PagesMultiprocessing Implementation

  At first I thought that this was goofy, but then I logged in and saw that Megatron only has 8GB of RAM. I guess that used to be cool in the old days, but now is tiny (my laptop has 8 GB of RAM). I'll see if someone around has some free RAM for a 4600; in the meantime, I've killed a MEDM that was running on there and using up a few hundred MB.

Run your ssh-MEDMs elsewhere or else I'll make a cronjob to kill them periodically.

  8210   Sat Mar 2 00:09:31 2013 ranaUpdateLSCXarm oscillation stopped

  Don't use resonant gain - it can lead to a loop instability since it makes the loop have 3 UGFs.

Just use a elliptic bandstop filter at this harmonic frequency separately for each test mass. There are many detailed examples of this in elog entries from Rob and I over the past ~10 years. This bandstop should get clicked on automatically after lock acquisition.

  8211   Sat Mar 2 00:23:19 2013 ranaSummaryCOCPhase Maps measured of the ATF flat mirrors

I took the two 'flat' 2" mirrors over to Downs and Garilynn showed me how to measure them with the old Wyko machine.

The files are now loaded onto our Dropbox folder - analysis in process. From eyeball, it seems as if the RoCs are in the neighborhood of ~5 km, with the local perturbations giving ~10-15 km of curvature depending upon position (few nm of sage over ~1 cm scales)

Koji's matlab code should be able to give somewhat more quantitative answers...

Ed: Here you are. "0966" looks good. It has RoC of ~4km. "0997" has a big structure at the middle. The bump is 10nmPV (KA)

 

Attachment 1: 0966_0997_phasemap.pdf
0966_0997_phasemap.pdf 0966_0997_phasemap.pdf 0966_0997_phasemap.pdf 0966_0997_phasemap.pdf 0966_0997_phasemap.pdf 0966_0997_phasemap.pdf 0966_0997_phasemap.pdf 0966_0997_phasemap.pdf
  8213   Sat Mar 2 14:52:02 2013 ranaUpdateLSCstable lock of PRMI

  Whereas the sensing matrix coefficients for ITMX/ITMY in REFL_I indicate an actuation imbalance, the disparity in the ITMX/ITMY to AS_Q elements does not. However, they do indicate why there is a PRM to AS_Q coupling at all.

I would recommend setting up the triggering so that the REFL & AS whitening is turned on AFTER lock acquisition and using a frequency of ~100-300 Hz for the sensing matrix measurement to fix these issues.

  8217   Mon Mar 4 09:55:33 2013 ranaUpdateLSCLSC whitening triggering started

  How about posting a logic flow diagram? Is the idea to trigger only on the power signals to determine the lock state? Is the hysteresis going to be done in the same way as the main filter bank triggers?

  8326   Thu Mar 21 12:33:51 2013 ranaUpdateComputersFixed

Please stop power cycling computers. This is not an acceptable operation (as Jamie already wrote before). When you don't know what to do besides power cycling the computer, just stop and do something else or call someone who knows more. Every time you kill the power to a computer you are taking a chance on damaging it or corrupting some hard drive.

In this case, the right thing to do would be to hook up the external keyboard and monitor directly to c1sus to diagnose things.

NO MORE TOUCHING THE POWER BUTTON.

  8337   Mon Mar 25 15:24:41 2013 ranaUpdateLockingTT1 drift

 

 I doubt that it has truly disappeared. Can you please make a measurement of it and quantify the hysteresis using some angle sensor?

  8343   Mon Mar 25 23:17:11 2013 ranaSummaryIOOMC

I measured the dark noise and regular intensity noise in MC trans tonight to get some estimate for the free running spectrum that the Chas ISS must overcome. PDF is attached.

XML file is /users/rana/dtt/MC-trans_130325.xml

The RIN normalization was done by using the mean of the SUM time series. The dark noise was measured by misaligning MC2 and making the same measurement with the bright normalization.

Attachment 1: MC-trans.pdf
MC-trans.pdf
  8354   Tue Mar 26 15:43:45 2013 ranaUpdateIOOTuning MC length/FSR

  Please attach the data file of the measurement - its hard to get the real information out of picture. In general, WE should always include the code / explanation of how to reconstruct the plot and get the scientific information out.

  8364   Thu Mar 28 04:21:40 2013 ranaUpdateLockingPRMI progress - even better!

 

 A key step was turning off the whitening filters. With the previous setting (G = 15 dB, white on), the error signals (post anti-whitening) had amplitudes of ~500 counts. This means that they can go as high as (150/15)^2 * 500 = 50000 counts on the ADC.

The purpose of the whitening filter is to match the noise / range of the signal to the ADC. What we would like to do is use the minimum gain so as to make the RFPD electronics noise + shot noise be ~equal to the ADC noise. i.e.

sqrt(V_PD^2 + v_shot^2) * G_white = V_ADC

The RFPD noise is ~3 nV before the internal preamp. The MAX4107 has a gain of 10. There is a factor of 1/2 from the voltage division of the RFPD's 50 Ohm series resistor and the input impedance of the mixer. There is also a power splitter between the PD output and the mixer which gives us a 3 dB loss. The mixer has a conversion loss of ~5-6 dB depending upon the LO level.

V_PD = 3e-9 * (10 * 1/2 * 1/sqrt(2) * 1/2) = 5e-9 V/rHz   (this is already bad; the signal coming out of the mixer needs to be amplified by x10 before going out to the whitening board).

In any case, its clear that we need something like 60 dB of gain for the PD noise to match the ADC noise. This is why increasing the whitening gain improves the error signal's SNR, reduces the hash driving the optics, and improves the locking. We should run with 45 dB gain and the switch on whitening after the lock.

 

Even better would be to modify the LT1128 input stage of the card to have the single stage of fixed whitening as we did for iLIGO. Then we can have triple whitening in lock.

  8378   Sun Mar 31 17:26:32 2013 ranaUpdateLSCLock of PRMI on sidebands

 

 Our first move has to be fixing the whitening switching for REFL55. That's the configuration we need to start and then move onto REFL165 to get to FPPRMI.

  8390   Tue Apr 2 16:13:10 2013 ranaUpdateSUSoplev calibration for ITMX, BS

  Very good - now you need to just put the cal factor into the filter banks so that the PERROR and YERROR signals are in microradians all the time.

 

EDIT JCD:  In progress.

  8394   Tue Apr 2 20:52:35 2013 ranaUpdateComputersiMac bashed

 I changed the default shell on our control room iMac to bash. Since we're really, really using bash as the shell for LIGO, we might as well get used to it. As we do this for the workstations, some things will fail, but we can adopt Jamie's private .bashrc to get started and then fix it up later.

  8397   Tue Apr 2 23:14:02 2013 ranaUpdateLSC22/110MHz path for POP

Quote:

We could not find a power supply slot for the amplifiers on the LSC rack. We had to put a temporary power supply in contradiction to our 'no temporary power supply' policy.

 After 1 month, its hard to imagine that this could not have been fixed by putting in a proper fuse and fuse block. I will remove this tomorrow if I still find it this way in the bottom of the rack.

There are also 2 Sorensen switching supplies in the bottom of the LSC rack (with all of our sensitive demod boards). These should also be moved over to the old 'digital' LSC rack tomorrow for the post meeting lab cleanup.

Use fuse blocks with fuses with appropriate ampacity.

  8409   Wed Apr 3 22:26:51 2013 ranaConfigurationElectronicsputting together a 110 MHz LSC demod board

 For the 110 MHz demod boards, we would ideally have a plugin bandpass filter. If you have some specs in mind, you can email mini-circuits or pulsar microwave about making a custom part; its not too expensive usually.

For the meantime, you should remove the onboard one and replace with a combination of low/high pass filters from Mini-Circuits. If you put a SLP-150 and a SHP-100 in series, the insertion loss should be less than 1 dB.

I think the ERA amps are OK for now, but they die with time, so they just need to be tested and replaced if necessary.

  8410   Wed Apr 3 23:22:20 2013 ranaUpdateoptical tablesOptical Table Toolboxes Update

 There are some tips for how to appy nail polish on YouTube from MKNails and MissJenFABULOUS. Their tips on how to prepare the site for a strong bonding strength are probably helpful for our gold/nickel coated tools. For chrome tools we may need to abrade the surface with a stone or fine sandpaper for it to take the layer better. IF the YouTube videos don't do it for you, then I suggest contacting Tom Evans at LLO to find out what kind of nail polish he uses.

  8419   Sat Apr 6 09:21:36 2013 ranaUpdate40m UpgradingNot a fan of the new plastic box yet

 

 1) We still need to drill and install the thumbscrew latches which secure the lids to the table. We cannot use the tables as an acoustic enclosure with loose lids.

2) For the camera issue, the idea is to put the longpass filters on the front of the cameras: then they are only sensitive to light with wavelength > 800 nm.

3) Whenever any interferometer work is happening the light switches must be in the positions which have been marked on them (and which most everyone ignores foolishly). We have never been insensitive to the room lights; the black table enclosures just give us a false sense of security. Room lights impact the interferometer noise.

  8424   Mon Apr 8 22:43:44 2013 ranaUpdatePSLMC locking troubles: MC/FSS servo unstable

The MC seemed to be losing lock recently quite a bit. I noticed that the PC Drive RMS signal was red.

This means that the high frequency drive to the Pockels cell was too high by a factor of 2-3 and sometimes saturating and breaking the lock.

I fiddled with the gains on the FSS screen until this value went down. It looks like there is some kind of high Q oscillation; it takes a couple minutes for the PC Drive RMS to settle to its new position after changing the gains.

The attached trend plot show the last 2 hours. The mean is now back to ~1 V and seems OK. We should really examine the FSS or MC error point spectra with the RF analyzer while exploring this gain space.

Attachment 1: Untitled.png
Untitled.png
  8426   Tue Apr 9 00:32:39 2013 ranaConfigurationIOOTurn on MCL

 Listening to the free swinging Y-arm, its clear that the fringe velocity is higher with the MCL path off, than on. I checked that the default gain of -300 was too high and changed it to -100 in the mcup script. With the higher gain value, there was clear gain peaking at just under 60 Hz (where the 60 Hz comb filter starts). We basically want the UGF to be between 20 and 60 Hz so that we can have the Bounce mode RG at 16.25 Hz and also the notch filter at 60 Hz.

Den is measuring and setting the UGF to be ~45 Hz.

With the MCL path on there is a high frequency horn-like noise in the Yarm when it locks. Its reduced a little by reducing the loop gain, but clearly this is just noise introduced by the MCL loop as Den noted before (and also tonight). He is cleaning up the whole MCL MEDM situation so that its useable by us. At the moment I've re-enabled it in the mcup.

My belief is that the frequency noise from the unstabilized MC is making the PRC locking harder. This will be investigated by tuning the shape of the MCL/MCF crossover so that we can turn it on without ruining the arm cavity spectra. Since the PRC length is ~2x smaller than the MC, we would expect it to be less sensitive to the MC frequency noise. But, since there is some common mode rejection in there, this may not be true. We'll only know by measuring PRC control signal with MCL on/off.

  8447   Fri Apr 12 09:20:32 2013 ranaUpdateLockingprcl angular motion

 How is the cavity g-factor accounted for in this calculation?

  8450   Sat Apr 13 03:45:51 2013 ranaUpdateLockingprcl angular motion

 

 Maybe its equivalent, but I would have assumed that the input beam is fixed and then calculate the cavity axis rotation and translation. If its small, then the modal expansion is OK. Otherwise, the overlap integral can be used.

For the ETM motion, its a purely translation effect, whereas its tilt for the ITM. For the PRM, it is also a mostly translation effect as calculated at the PRC waist position (ITM face).

  8454   Sun Apr 14 17:56:03 2013 ranaUpdateLockingprcl angular motion

Quote:

Quote:

For the PRM, it is also a mostly translation effect as calculated at the PRC waist position (ITM face).

I made another estimation assuming that PRCL RIN is caused by translation of the cavity axis:

  • calibrated RIN to translation, beam waist = 4mm

 In order to get translation to RIN, we need to know the offset of the input beam from the cavity axis...

This should be possible to calibrate by putting a pitch and yaw excitation lines into the PRM and measuring the RIN.

See secret document from Koji.

  8471   Mon Apr 22 17:06:42 2013 ranaSummaryIOOMC locked/aligned. MC WFS offloading by ezcaservo

 

 Why use the PSL beam as a reference? Don't we want to keep the MC pointing in a good direction through the Faraday instead???

  8479   Tue Apr 23 22:10:54 2013 ranaUpdateComputersNancy

controls@rosalba:/users/rana/docs 0$ svn resolve --accept working nancy
Resolved conflicted state of 'nancy'

  8480   Tue Apr 23 22:59:05 2013 ranaConfigurationSUSOptical Lever Gains normalized

Due to the recent addition of cal factors in the OL error points, the OLPIT_GAIN and OLYAW_GAIN have been reduce to tiny numbers (e.g. 0.002).

Since our MEDM only shows 3 digits past the decimal point by default, it makes more sense to have the gains around 1.

So I reduced the gains in all of the FM1 filters from 1000 to 1 and multiplied the GAIN values by 1000 (using ezcastep) to compensate.

All of the active optics seem to be behaving as before. Haven't tested ETMs or SRM yet.

  8529   Sat May 4 00:21:00 2013 ranaConfigurationComputersworkstation updates

 Koji and I went into "Update Manager" on several of the Ubuntu workstations and unselected the "check for updates" button. This is to prevent the machines from asking to be upgraded so frequently - I am concerned that someone might be tempted to upgrade the workstations to Ubuntu 12.

We didn't catch them all, so please take a moment to check that this is the case on all the laptops you are using and make it so. We can then apply the updates in a controlled manner once every few months.

  8544   Tue May 7 19:58:28 2013 ranaFrogsTreasurerabbitt whole

controls@rosalba:/opt/rtcds/caltech/c1/medm/c1ioo/master/master/master 0$ ls
C1IOO_LKIN_OUT_MTRX.adl   C1IOO_MC_ASS_LOCKIN5.adl     C1IOO_WFS1_I.adl             C1IOO_WFS_LKIN.adl
C1IOO_LOCKMC.adl          C1IOO_MC_ASS_LOCKIN6.adl     C1IOO_WFS1_Q.adl             C1IOO_WFS_MASTER.adl
C1IOO_LOCKMC_BAK.adl      C1IOO_MC_ASS_PIT_LOCKIN.adl  C1IOO_WFS1_SETTINGS.adl      C1IOO_WFS_MASTER.adl~
C1IOO_MC_ALIGN.adl        C1IOO_MC_ASS_YAW_LOCKIN.adl  C1IOO_WFS1_SETTINGS.adl.old  C1IOO_WFS_MASTER_BAK.adl
C1IOO_MC_ALIGN.adl~       C1IOO_MC_LOCKINS.adl         C1IOO_WFS2_I.adl             C1IOO_WFS_OUTMATRIX.adl
C1IOO_MC_ALIGN_BAK.adl    C1IOO_MC_SERVO.adl           C1IOO_WFS2_Q.adl             C1IOO_WFS_QPD.adl
C1IOO_MC_ASS.adl          C1IOO_MC_TRANS_QPD.adl       C1IOO_WFS2_SETTINGS.adl      C1IOO_WFS_QPD.adl.old
C1IOO_MC_ASS_LOCKIN1.adl  C1IOO_Mech_Shutters.adl      C1IOO_WFS2_SETTINGS.adl.old  fmX
C1IOO_MC_ASS_LOCKIN2.adl  C1IOO_MODECLEANER.adl        C1IOO_WFS_HEADS.adl          junk
C1IOO_MC_ASS_LOCKIN3.adl  C1IOO_QPDS.adl               C1IOO_WFS_HEADS.adl.old      master
C1IOO_MC_ASS_LOCKIN4.adl  C1IOO_QPDS_BAK.adl           C1IOO_WFS_INMATRIX.adl       svn-commit.tmp~
controls@rosalba:/opt/rtcds/caltech/c1/medm/c1ioo/master/master/master 0$ cd master
controls@rosalba:/opt/rtcds/caltech/c1/medm/c1ioo/master/master/master/master 0$ cd master
controls@rosalba:/opt/rtcds/caltech/c1/medm/c1ioo/master/master/master/master/master 0$ cd master
controls@rosalba:/opt/rtcds/caltech/c1/medm/c1ioo/master/master/master/master/master/master 0$ cd master
controls@rosalba:/opt/rtcds/caltech/c1/medm/c1ioo/master/master/master/master/master/master/master 0$ cd master
controls@rosalba:/opt/rtcds/caltech/c1/medm/c1ioo/master/master/master/master/master/master/master/master 0$ cd master
controls@rosalba:/opt/rtcds/caltech/c1/medm/c1ioo/master/master/master/master/master/master/master/master/master 0$ cd master
controls@rosalba:/opt/rtcds/caltech/c1/medm/c1ioo/master/master/master/master/master/master/master/master/master/master 0$ cd master
controls@rosalba:/opt/rtcds/caltech/c1/medm/c1ioo/master/master/master/master/master/master/master/master/master/master/master 0$ cd master
controls@rosalba:/opt/rtcds/caltech/c1/medm/c1ioo/master/master/master/master/master/master/master/master/master/master/master/master 0$ cd master
controls@rosalba:/opt/rtcds/caltech/c1/medm/c1ioo/master/master/master/master/master/master/master/master/master/master/master/master/master 0$ cd master
controls@rosalba:/opt/rtcds/caltech/c1/medm/c1ioo/master/master/master/master/master/master/master/master/master/master/master/master/master/master 0$ cd master
controls@rosalba:/opt/rtcds/caltech/c1/medm/c1ioo/master/master/master/master/master/master/master/master/master/master/master/master/master/master/master 0$ cd master
controls@rosalba:/opt/rtcds/caltech/c1/medm/c1ioo/master/master/master/master/master/master/master/master/master/master/master/master/master/master/master/master 0$ cd master
controls@rosalba:/opt/rtcds/caltech/c1/medm/c1ioo/master/master/master/master/master/master/master/master/master/master/master/master/master/master/master/master/master 0$ cd master
controls@rosalba:/opt/rtcds/caltech/c1/medm/c1ioo/master/master/master/master/master/master/master/master/master/master/master/master/master/master/master/master/master/master 0$ cd master
controls@rosalba:/opt/rtcds/caltech/c1/medm/c1ioo/master/master/master/master/master/master/master/master/master/master/master/master/master/master/master/master/master/master/master 0$ cd master
controls@rosalba:/opt/rtcds/caltech/c1/medm/c1ioo/master/master/master/master/master/master/master/master/master/master/master/master/master/master/master/master/master/master/master/master 0$ cd master
controls@rosalba:/opt/rtcds/userapps/trunk/isc/c1/medm/c1ioo 0$ cd master
controls@rosalba:/opt/rtcds/userapps/trunk/isc/c1/medm/c1ioo/master 0$ cd master
controls@rosalba:/opt/rtcds/userapps/trunk/isc/c1/medm/c1ioo/master/master 0$ cd master
controls@rosalba:/opt/rtcds/userapps/trunk/isc/c1/medm/c1ioo/master/master/master 0$ cd master
controls@rosalba:/opt/rtcds/userapps/trunk/isc/c1/medm/c1ioo/master/master/master/master 0$ helppp
helppp: command not found
controls@rosalba:/opt/rtcds/userapps/trunk/isc/c1/medm/c1ioo/master/master/master/master 127$ help me
bash: help: no help topics match `me'.  Try `help help' or `man -k me' or `info me'.

  8546   Tue May 7 20:35:15 2013 ranaUpdatePSLPSL database / screen maintenance
  1. Fixed LOCKMC screen to display the SLOWM voltage instead of the ref cav transmission.
  2. PSL Status .db file updated to trigger on correct limits for FSS FAST and the FSS and PMC local oscillator levels.
  3. SETTINGS_SET screen can now take and display screen snapshots.

All of these changes were committed to the SVN.

  8577   Wed May 15 00:45:28 2013 ranaConfigurationLSCOpenloop gain for PRMI lock May 13

 

 Pfft. Why 500 usec delay? We should be using the known parameters for the hardware and software AA/AI.

  8583   Wed May 15 19:32:04 2013 ranaSummaryCDSAccounting of ADC/DAC channel availability
  1. What are we using 16 DAC channels for in the LSC?
  2. What are the functions of those IOO DAC channels which go to cross-connects? If they're not properly sending, then we may have malfunctioning MC or MCWFS.
  3. Can we just use the SLOW DAC (4116) for the ALS PZTs? We used this for a long time for the input steering and it was OK (but not perfect).
  8586   Wed May 15 23:38:04 2013 ranaUpdateelogcategories trimmed

I deleted a bunch of useless categories from the 40m elogd.cfg file. IF you find your category has been deleted, you can now learn how to categorize yourself into the existing categories. ELOG restarted and log file zeroed.

  8588   Thu May 16 02:34:38 2013 ranaUpdateComputer Scripts / ProgramsAutoRUN GUI resurrected

We talked about the thing that watches the scripts for autolocking during the meeting today.

I've resurrected the Perl-Tk GUI that we used through i/eLIGO for watching the IFO and running the appropriate scripts. This is not meant to be a replacement for aLIGO stuff, but just something to get us going for now. I expect that we will make some new fanciness which will eclipse this, but I brought it back so that we don't start off with some 'Advanced' system which is worse than the old one.

You can run it from scripts/c1/ by typing ./AutoRUN.pl. It pops up the GUI and starts in a Disabled mode where it watches and does nothing.

I have done some editing of the GUI's code so that it uses caget / caput instead of ezca binaries. New stuff is in the SVN.

Next up is to start testing it and fixing it up so that it uses the thresholds set in the LSC screens rather than some hardcoded values.

Eventually we should also convert all of its daughter scripts from tcsh to bash to keep Jamie's blood pressure in the low hundreds...

Attachment 1: Autorun.png
Autorun.png
  8596   Fri May 17 16:06:14 2013 ranaUpdate40m upgradingETMY op table disabled

 

 TODO:

  1. We need to replace all of the floppy anodized Al dumps with clean razor blade dumps on stiff mounts. BOTH of the rejected ports of the 1064 FI need some kind of custom dump.
  2. All of the leakthrough beams of the HR mirrors also need razor dumps. A good rule of thumb is that your first notion of how to implement the beam dump is NOT good enough.
  3. The whole lens / modematching situation for the 1064 and 532 paths will have to be redone so as to put the beam waists inside the Faraday crystal (NOT outside). The beam waist in the 4.7 mm diameter Faraday should be ~0.3-0.4 mm.
  4. The efficiency that we got for the doubling shows that we don't need the cylindrical lenses - they are nice, but not needed to get 90% of the max power.
  5. The lenses between the 1064 FI and the doubler should be put onto a base that can be used to adjust the lens position for MM optimization. Nothing fancy, just something slideable.
  6. For this iteration of the table, we can do as Annalisa has written so as to get the green MM lenses ordered ASAP. After next week we can come up with a new plan before dismantling the EX table.
  8667   Mon Jun 3 22:36:07 2013 ranaUpdateLSCALS-TRY_OUT needs anti-whitening

 After working some more on the EY table, we are getting some TEM00 flashes for the Y arm green. We have had to raise the height of one of the MM lenses to prevent clipping.

We used a function generator to apply a ~300 mV 10 Hz triangle wave to scan the laser frequency while aligning.

We tried to use the C1:ALS-TRY_OUT channel to help us in our alignment but there are a couple problems:

1) It seems that there is an uncompensated whitening filter before the ADC - Annalisa is making a compensation filter now.

2) The data delay is too much to use this for fast alignment. We might need to get a coax cable down there or mount a wired ethernet computer on the wall.

3) We need to make DQ channels for the TRY and TRX OUT. We need long term data of these, not just test points.

Attachment 1: try.pdf
try.pdf
  8668   Tue Jun 4 10:33:09 2013 ranaUpdateSUSETMY oplev path redone

 We also redid the oplev path.

Someone had used the illegal tactic of using aluminum dogs to hold down mounts which already have screw holes in the bases. This is too flimsy to use.

We then adjusted the position of the second lens to get a ~1 mm spot on the QPD. There is a lot of stray red light from extra reflections, but so be it.

Its possible that we are now clipping the IPANG path, but we will need to lock the Y arm and verify the input pointing to be sure. At this point we are ready to measure the ETMY OL loop gain and tune, etc.

  8673   Tue Jun 4 20:19:07 2013 ranaConfigurationIOOChanged threshold for FSS SLOW loop

The FSS SLOW actuator often runs off away from zero and into a region where the mode hopping is bad and makes a lot of frequency noise (e.g. that 8 hour period a few weeks ago when Jamie couldn't lock the MC).

It should not have this behavior. The SLOW loop should only be running when the MC is locked.

Today I found that the threshold was set back to 0.2 V (which is approximately the correct value for the RefCav locking). Its being compared to the MC TRANS, so the correct value should be ~1/2 of the maximum MC TRANS.

To find this out, I read this piece of text:

    # Make sure the loop is supposed to be active
    if (get_value("C1:IOO-MC_TRANS_SUM") < get_value("C1:PSL-FSS_LOCKEDLEVEL")) {
    print("Reference Cavity not locked -- control loop disabled.\n");
    next;
    }

from scripts/PSL/FSS/FSSSlowServo. I set the threshold using the commmand:

caput -c -t C1:PSL-FSS_LOCKEDLEVEL 12000

Then I restarted the code on op340m, by typing:

> nohup FSSSlowServo

and then closing the terminal. Seems to be behaving correctly now. Previously, the value was so low that the SLOW loop was never turning itself off.

  8677   Wed Jun 5 14:01:37 2013 ranaUpdateLSCC1:ALS-TRY_OUT filter and green progress

The rejected beam from this Faraday comes out at a tiny, tiny angle and so its tough to pick it off without clipping the main beam.

Some care must be taken in setting this up - Steve may have some good ideas on what kind of mount can be placed so close to the beam.

Why did we ever order this terrible Faraday? Let's never get a Faraday with a tiny angle between the beams again.

  8683   Wed Jun 5 17:37:10 2013 ranaUpdateGeneralfast 2004 qpd

Quote:

Thank you Ben Abbott forwarding this information:

 QPD Amplifier D990272 https://dcc.ligo.org/cgi-bin/private/DocDB/ShowDocument?.submit=Number&docid=D990272&version= at the X-end.  It plugs into a Generic QPD Interface, D990692, https://dcc.ligo.org/cgi-bin/private/DocDB/ShowDocument?.submit=Number&docid=D990692&version= according to my drawings, that should be in 1x4-2-2A.

 Wrong: this is not an interface.