ID |
Date |
Author |
Type |
Category |
Subject |
4803
|
Fri Jun 10 12:02:10 2011 |
rana | Update | CDS | Second trends only go back 12 days |
Quote: |
While answering a quick question by Kiwamu, I noticed we only had second trends going back to 99050000 GPS time, May 27th 2011.
Trends (I thought) were intended to be kept forever, and certainly longer than full data, which currently goes back several months.
Jamie will need to look into this.
|
Our concept is to keep second trends for 1-2 months and minute trends forever. The scheme that Alan had worked out many years ago had it such that we could look back to 1998 and that the minute trends would be backed up somehow.
If its not working, we need to get Alan's help to recover the previous configuration. |
4830
|
Fri Jun 17 00:17:26 2011 |
rana | Configuration | Electronics | 2 RFPDs sent to LLO |
Koji and I found 2 RFPD boxes to send to LLO. We've put them onto Steve's desk to be overnighted to Valera.
One of them is our old 21.5 MHz gold box RFPD from the FSS (which we don't use). The other one is a 2mm gold box one which was previously tuned for 66 MHz.
They shipped out on Friday |
Attachment 1: P1070895.JPG
|
|
4843
|
Mon Jun 20 17:58:00 2011 |
rana | Update | CDS | Gateway program killed |
There was a rogue, undocumented, gateway process running on NODUS since ~4 PM. This guy was broadcasting channels back into the Martian and causing lockups in the IOO controls. I did a kill -9 on its process.
Someone will pay for this. |
4861
|
Wed Jun 22 21:36:41 2011 |
rana | Summary | General | July 2011 vent plan |
I put a paper Peet's bag with half of the Mini-Moos into George. |
4864
|
Thu Jun 23 09:46:16 2011 |
rana | Update | LSC | PRMI locking : not stable enough |
All the suspensions are bad until you fix them. But, ... there is a script which can be used to diagnose them today:
Python SUStest |
4881
|
Fri Jun 24 22:35:23 2011 |
rana | Configuration | CDS | dataviewer broken on pianosa |
When I try to get minute trend, it says "word too long".  |
4886
|
Sun Jun 26 16:17:22 2011 |
rana | Update | CDS | diagonalization of MC input matrix |
I have updated the scripts/SUS/peakFit/ directory so that it now finds the SUS input matrix coefficients in addition to just finding the free-swinging peaks.
Procedure:
- Get OSEM sensors data via NDS2 from a time when the optics have been kicked and then left free swinging.
- Downsample the data to 64 Hz and save.
- Make power spectra with a 1 mHz resolution (i.e. we need a few hours of data) and ~10 averages.
- use the fminsearch lorentzian peak fitter -> save the peak frequencies
- Make Transfer Function estimate matrix at the peak frequencies between all OSEMs (this makes a 5x4 complex matrix)
- The matrix should be real, so make sure its mostly real and then take the real part only
- Normalize (height of biggest peak for each f_DOF should be 1)
- Add a Butterfly mode vector. This makes the sensing matrix go from 5x4 to 5x5. (Butterfly a.k.a. Pringle)
- Invert
- Normalize so that the biggest element in each Sensor2DOF column is 1.
- Load values into MEDM screen and then verify by another free swinging data run.
The attached PDF shows how much rejection of the unwanted DOFs we get between the existing diagonal input matrix and this new empirical matrix. Previously, the decoupling was only a factor of a few for some of the modes. Now the decoupling is more like orders of magnitude (at least according to this calculation). It will be worse when we load it and then try another free swinging run. However, the fact that the suppression can be this good means that the variation in the coefficients at the ~hours time scale is at least this small (~< 0.1%)
That's the basic procedure, but there are a lot of important but mainly technical details:
- Free swinging data must be taken with the angle bias ON. Otherwise, we are not measuring the correct sensing gain (i.e. the magnets are not in their nominal place within the OSEM-LED beam)
- Data must be checked so that the shadow sensor outputs are in their linear regime: if they are exploring the cubic part, then the fundamental is being suppressed.
- Instead of just using the peak frequency, I average a few points around the peak to get better SNR before inversion. I think this will make the results more stable.
- All previous input matrix diagonalization efforts (Buckley, Sakata & Kawamura, Black, Barton, Gonzalez, Adhikari & Lawrence, Saulson,...) for the past ~15 years have been using the spectra's peak height data. Today's technique uses the TF and so is more precise. The coherent transfer function is always better than just using the magnitude data.
- This method is now fairly automatic - there's no human intervention in fudging values, choosing peak heights, frequencies, etc.
- We'll have to rerun this, of course, after the mirrors are aligned and after the OSEM whitening fiasco is cleaned up somewhat.
I'll set the optics to be aligned and then swing tonight. |
Attachment 1: inMatDiag.pdf
|
|
4887
|
Sun Jun 26 18:35:16 2011 |
rana | HowTo | SUS | free swing all optics |
I used scripts/SUS/freeswing-all.csh to give the optics a kick and then turn off their watchdogs and collect the free swinging data. Final script end time = 993173551. Start taking data ~ 993173751
I had to fix up the script a little: it had amateur stuff in there, such as undefined variables.
It still doesn't work that well. On the new Ubuntu workstations, pianosa, it fails by just not setting some of the EPICS variables using the EZCA stuff.
On Allegra, it failed on ~1 out of 10 commands by returning "epicsThreadOnce0sd epicsMutexLock failed" ???
On Pianosa, it sometimes says, instead, "epicsThreadOnceOsd: pthread_mutex_lock returned Invalid argument. ". Ah...now I understand?
So finally, I had to run the script on op340m to get it to actually run all of its commands. That's right; I used a 15 year old Solaris 9 Blade 150 because none of our fancy new Linux machines could do the job reliably.
Fixing our EZCA situation is a pretty high priority; if the locking scripts fail to run ~1 command every hour its going to completely derail the lock acquisition attempts.
If you want to use the IFO tonight, just run the script again on op340m again when you're done. |
Attachment 1: ringdown.png
|
|
4888
|
Sun Jun 26 22:38:20 2011 |
rana | Update | CDS | MC1 LR dead for > 1 month; now revived temporarily |
Since the MC1 LRSEN channel is not wasn't working, my input matrix diagonalization hasn't worked today wasn't working. So I decided to fix it somehow.
I went to the rack and traced the signal: first at the LEMO monitor on the whitening card, secondly at the 4-pin LEMO cable which goes into the AA chassis.
The signal existed at the input to the AA chassis but not in the screen. So I pressed the jumper wire (used to be AA filter) down for the channel corresponding to the MC1 LRSEN channel.
It now has come back and looks like the other sensors. As you can see from this plot and Joe's entry from a couple weeks ago, this channel has been dead since May 17th.
The ELOG reveals that Kiwamu caught Steve doing some (un-elogged) fooling around there. Burnt Toast -> Steve.

993190663 = free swinging ringdown restarted again |
Attachment 1: lrsen.png
|
|
4889
|
Mon Jun 27 00:23:11 2011 |
rana | Update | CDS | ETMX SIDE problem |
The slow readback of the ETMX side seems to also have something flaky and bi-stable. This is not an issue for damping, but it disables the SIDE watchdog for ETMX and makes it unsafe if we accidentally use the wrong damping sign. |
Attachment 1: etmx-side.png
|
|
4892
|
Tue Jun 28 01:18:53 2011 |
rana | HowTo | SUS | free swing all optics |
Chris Wipf tells me that the EPICS Mutex Jumbo Mumbo can be overcome by upgrading our EPICS. We should get one of Jamie's assistants to get this going on one of the Ubuntu workstations. |
4914
|
Wed Jun 29 22:50:02 2011 |
rana | Update | IOO | misc. MC work |
Today I wanted to investigate the MC Length path situation for obscure reasons.
Jamie has started to revert the "ALTPOS" effect on the MC mirrors. So far, the screens and SLOW channels have been fixed, but the fast channels still say "ALTPOS" in the dataviewer instead of "MCL".
I also noticed that all of our old ADCU channels for diagnosing the PSL, MC, ISS, PMC ,etc. are completely AWOL. Let's blame Joe.
I think that there are probably some ADC channels available and that we'll just have to figure out what Joe intended for this. We certainly need it if we want to diagnose our PMC, ISS, FSS, MC, etc. Kiwamu tells us that the old PSL/IOO AA chassis is being used for some of the GCV signals, so its likely that we just have to do the appropriate channel name mapping in the DAQCONFIG tool.
Forging ahead with no data, I made up some filters in the MC2-MCL filter bank so that there could be a stable crossover between the laser path. I was able to turn it on and get some suppression of the FSS-FAST control signal, but there's no way to be sure without the fast channels. We gotta get Jamie to help us out once he finished the ETM BO mess. |
4926
|
Thu Jun 30 21:55:16 2011 |
rana | Configuration | DAQ | NDS2 conf change |
As I recently had trouble getting all of the SUS SENSOR channels at once from NDS2, I asked J.Z. for help. He found that the number of buffers on mafalda was set to only allow a small amount of data to be requested at one time.
He's going to have to figure out a more permanent fix, but for now he's increased the data buffer size to allow somewhat larger chunks to be gotten. I have made a work around in matlab, which gets smaller chunks and then cats them together.
Its in SUS/peakFit/. |
Attachment 1: Untitled.png
|
|
4928
|
Fri Jul 1 11:47:25 2011 |
rana | Update | IOO | WFS2 resonances and installation |
What is implicit in Suresh's entry is that we decided to run the WFS with the 10 dB internal attenuation set to ON as the nominal. In the past, we have always had all the attenuation OFF for max gain. The layout of the WFS is such that we get that nasty 200 MHz oscillation due to crosstalk between the 2 MAX4106 opamps for each quadrant. The 10 dB attenuator is able to reduce the positive feedback enough to damp the oscillation.
In principle, this is still OK noise-wise. I think the thermal noise of the resonant circuit should be ~2-3 nV/rHz. Then the first opamp has a gain of 5, then the -10 dB attenuator, then another gain of 5. The noise going to the demod board is then ~10-15 nV.
The real noise issue will be the input noise of the demod board. As you may recall, the output of the AD831 mixer goes to a AD797. The AD797 is a poor choice for this application. It has low noise only at high frequencies. At 10 Hz, it has an input voltage noise of 10 nV/rHz and a current noise of 20 pA/rHz. If we wanted to use the AD797 here, at least the RC filter's resistor should be reduced to ~500 Ohms. Much better is to use an OP27 and then choose the R so as to optimize the noise.
We should also be careful to keep the filter frequency low enough so as not to rate limit the OP27. From the schematic, you can see that this circuit is also missing the 50 Ohm termination on the output. There ought to be the usual high-order LC low pass at the mixer output. The simple RC is just not good enough for this application.
As a quick fix, I recommend that when we next want to up the WFS SNR, we just replace the RC with an RLC (R = 500 Ohms, L = 22 uH, C = 1 uF).
|
Attachment 1: Screen_shot_2011-07-01_at_11.13.01_AM.png
|
|
4933
|
Fri Jul 1 20:22:24 2011 |
rana | Update | SUS | ETMY sus controller found to be in a bad state |
Actually, ETMY was the only good one. They should all have the 30 Hz High pass as the damping filter. I think these details are in the elog entry that we originally made while doing ETMY.
They should all also have a 3:30 in the XXSEN to compensate the whitening. The logic is supposed to be that FM1 is ON when the hardware whitening is ON. This is the opposite of the old logic and its why the damping filter has to be moved from 3 to 30 Hz. |
4934
|
Fri Jul 1 20:26:29 2011 |
rana | Summary | SUS | All SUS Peaks have been fit |
MC1 MC2 MC3 ETMX ETMY ITMX ITMY PRM SRM BS mean std
Pitch 0.671 0.747 0.762 0.909 0.859 0.513 0.601 0.610 0.566 0.747 0.698 0.129
Yaw 0.807 0.819 0.846 0.828 0.894 0.832 0.856 0.832 0.808 0.792 0.831 0.029
Pos 0.968 0.970 0.980 1.038 0.983 0.967 0.988 0.999 0.962 0.958 0.981 0.024
Side 0.995 0.993 0.971 0.951 1.016 0.986 1.004 0.993 0.973 0.995 0.988 0.019
There is a large amount of variation in the frequencies, even though the suspensions are nominally all the same. I leave it to the suspension makers to ponder and explain. |
Attachment 1: Screen_shot_2011-07-01_at_8.17.22_PM.png
|
|
4935
|
Sun Jul 3 21:18:06 2011 |
rana | Update | Computer Scripts / Programs | statScreen scripts dead since Feb 4 / now revived |
This CSHRC mangling on Feb 4 did more than re-arrange FB binaries.
It broke the path to MEDM for the 32-bit machines in the lab (e.g. mafalda) and stopped the MEDM snapshots from being posted onto our MEDM Status Web Page.
This is because, in addition to the paths mentioned in the above elog, the paths to the EPICS directories were also commented out. I've re-inserted them into our
.cshrc file in the 32-bit section; the statScreen CRON that Yoichi set up is now back in business.
* for some reason, the 'cronjob.sh' script is wiping out its own log file. It would be great if someone who understands stderr output re-direction can fix it so that the log-file from each run is retained until the next time cron runs. |
4942
|
Tue Jul 5 21:26:51 2011 |
rana | Update | SUS | More normalization of all sus controllers |
This is getting closer, but with the whitening left OFF and the cts2um filter also OFF, none of the suspensions are working correctly. I'm shutting down all the watchdogs until someone gets around to setting the damping gains and filters correctly.
I'm attaching a screenshot of some of the problems I see so far with MC3.
I'm going to try to get the MC suspensions working OK for tonight so that we can use them for the PRMI locking work.
Update #1: None of the MC SUS DAQ channels are found by dataviewer....SUS debugging speed reduced by 10x. Tue Jul 05 21:38:17 2011
Update #2: POS/PIT/YAW BIAS sliders now seem to work, but are ~1000x too weak to do anything. Tue Jul 05 21:41:38 2011
|
Attachment 1: Screenshot-1.png
|
|
4972
|
Fri Jul 15 09:25:02 2011 |
rana | Update | SUS | SUS oplev spectras |
In addition to the OL quadrants, you need to plot the OPLEV_PERROR and OPLEV_YERROR signals since these are the real signals we use for finding the mirror motion. If they're not in the Dataviewer, Jamie should add them as 256 Hz DAQ channels (using these names so that we have the continuity with the past). These DAQ channels correspond to the IN1 channels for the OL filter banks.
Also JPG are banned from the elog - you should put all of the plots into a single, multipage PDF file in honor of the new Wagonga. |
4988
|
Tue Jul 19 10:18:24 2011 |
rana | Update | LSC | Big ol' mess |
Remember, as per our marker board conversation, that the resistor noise as excitation method only works if the gain of all of the channels is set to something high (like 45 dB).
At 0 dB, the resistor noise is only 30 nV/rHz, whereas the ADC noise is more like 10000 nV/rHz... |
4991
|
Tue Jul 19 20:36:08 2011 |
rana | Update | Computers | VirtualBox + Windows 7 on rossa |
I installed Virtual Box on rossa. Then I put Windows 7 in there and am now installing Altium.
You can run Windows on rossa by just clicking Applications -> System Tools -> Virtual Box. |
5004
|
Wed Jul 20 19:24:12 2011 |
rana | Update | SUS | oplev gains today |
I guess Valera forgot to elog it. Steve, please email him and get the info.
I started to check out the OL servos today so that our whole interferometer is not too floppy.
- The ETMX OPLEV DAQ channels were not in the list. Jamie ran the activateDQ.py script and it came back. Right now, we have no diagnostics to know if people have run this or not so the frames will have missing data now and again depending on how forgetful the rebooters are. Perhaps we can get activateDQ put into the make file???
- I turned ON all of the offset buttons on the OL1, etc. filter banks. This allows for the dark offsets to be set for the OL quadrants. With these buttons off it doesn't make any sense.
- I noticed that there are white (INVALID) fields all over the OPTLEV_SERVO screens. This is just because the new SUS models have not captured all of the functionality of the old system. Needs fixing.

Some of these OL spectra are not like the others...
 |
5023
|
Sun Jul 24 20:47:21 2011 |
rana | Summary | Electronics | AA filter tolerance analysis |
This is sort of OK, except the capacitor connects across the (+) terminals of the two input opamps, and does not connect to ground.
Also, we don't care about the CMRR at 64 kHz. We care about it at up to 10 kHz, but not above. The sample frequency of the ADC is 64 kHz, but all of the models run at 16 kHz or less, so the Nyquist frequency is 8 kHz.
And doesn't the value depend on the resistors? |
5025
|
Mon Jul 25 00:35:44 2011 |
rana | Update | SUS | something wrong with ETMY LR sensor |

Looks like either the LR OSEM is totally mis adjusted in its holder or the whitening eletronics are broken.
Also looks like the ETMY is just not damped at 1 Hz? How can this be?
I look at the SUS_SUMMARY screen which apparently only Steve and I look at:

Looks like the suspensions have factor of 10-100 different gains. Why?
** The ETMY just doesn't behave correctly when I bias it. Both pitch and yaw seem to make it do yaw. I leave this for Jamie to debug in the morning.
*** Also, the BIAS buttons are still broken - the HOPR/LOPR limits ought to be 5000 and the default slider increment be 100. Also the YAW button readback doesn't correctly show the state of the BIAS.
**** And.....we have also lost the DAQ channels that used to be associated with the _IN1 of the SUSPOS/PIT/YAW filter modules. Please put them back; our templates don't work without them. |
5058
|
Thu Jul 28 21:52:40 2011 |
rana | Update | Computers | another attempt to use pianosa |
- Pianosa doesn't cache the SVN pwds so you need to re-enter at each SVN up or commit. This is different from the rest of our workstations. We need to determine what behavior we want.
- Tried to use the netGPIB scripts:
pianosa:gpib 0> ./readSR785.csh rb2
rb2
netgpibdata.py: Command not found. |
5065
|
Sat Jul 30 02:47:43 2011 |
rana | Update | PSL | ABSL Laser crystal temp left largely excited & left unattended for more than 3hours |
Hmm. Should have only been +/- 1 GHz. Some setting got changed apparently...
This is a part of the RefCav temperature measurement setup. You'll get an elog from Jenny very soon. |
5072
|
Sat Jul 30 20:41:50 2011 |
rana | Update | PSL | RefCav Stabilization back on |

I turned the RefCav heater and servo back on a couple days ago. At first it was stabilizing again at a low setpoint, but in reality the right temperature (~40 C). After fixing the in-loop signal offsets, the setpoint now correctly reflects the actual temperature.
Jenny is going to calibrate the sensors using some kind of dunking cannister next week. |
5081
|
Mon Aug 1 11:46:56 2011 |
rana | Summary | LSC | Tolerance of Arm length = 2 cm |
wow. This Monte-Carlo matrix is one of the most advanced optical modeling things I have ever seen. We never had this for any of the interferometers before. |
5095
|
Tue Aug 2 16:55:21 2011 |
rana | Update | ABSL | Arm length measurement : cavity kick technique |
Quote: |
I made some attempts to measure the current length of the arm cavities by using the mass-kicking technique.
|
Why not just scan the Green laser to measure the arm lengths instead? The FSR of the arm is ~3.75 MHz and so all you have to do is lock the arm green and then sweep the PZT until the resonance is found at 3.75 MHz.

|
5118
|
Thu Aug 4 11:18:12 2011 |
rana | Update | SUS | sus at atm |
Remember that the Oplevs are not good references because of the temperature sensitivity. The week long trend shows lots of 24 hour fluctuations. |
5131
|
Sat Aug 6 13:38:02 2011 |
rana | Summary | General | Summary of today's in-vacuum work |
This OSEM placement is just the OPPOSITE of what the proper placement is.
Usually, we want to put them in so that the LED beam is vertical. This makes the OSEM immune to the optic's vertical mode.
The orientation with the horizontal LED beam makes the immunity to the side mode better, but may spoil the vertical.
In reality, neither of these assumptions is quite right. The LED beam doesn't come out straight. That's why Osamu and I found that we have to put in some custom orientations.
Also, the magnet gluings relative to the OSEM bracket centers are not perfectly aligned. So...I am saying that the OSEMs have to be oriented empirically to reduce the couplings which we want to reduce.
|
5136
|
Mon Aug 8 00:12:58 2011 |
rana | Update | CDS | diagonalization of MC input matrix |
I've finally completed the SUS/peakFit/ scripts which find the new input matrix for the SUS. MC1, MC2, MC3, and ITMX have been matrix'd.
I tried to do the BS, but it came out with very funny matrix elements. Also the BS is missing its DAQ channels again (JAMIE !) so we can't diagnose it with the free swinging method.
To continue, we have to get some good data and try this again. Right now there are some weird issues with a lot of the optics. I've also set the damping gains for the optics with the new matrices.
Ex.
new_matrix = findMatrix('ITMX')
writeSUSinmat('ITMX', new_matrix)
writeSUSinmat.m
this script writes the values to the MEDM input SUS matrix. To do the writing, I used the low level 'caput' command instead of ezcawrite since the ezca libraries are getting deprecated.
caput doesn't really have good diagnostics, so I use matlab to check the return status and then display to the terminal. You can just rerun it if it gives you an error.
A coupled of normalization notes:
1) The POS/PIT/YAW rows are scaled so that the mean of abs(FACE elements) = 1. Previously, I had the max element = 1.
2) The SIDE row is scaled so that the SIDE element = +1.
3) I then normalized the ROWS according to the geometrical factors that Jamie has calculated and almost put into the elog.
All these scripts have been added to the SVN. I've removed the large binary data files from the directory though. You can just rsync them in to your laptop if you want to run this stuff remotely. |
5137
|
Mon Aug 8 00:58:26 2011 |
rana | Update | CDS | diagonalization of MC input matrix |
Besides the purpose of correctly tuning the suspensions, my hidden goal in the input matrix diagonalization has been to figure out what the 'true' sensing noise of the OSEMs is so that we can accurately predict the noise impact on the OAF.
The attached plot shows the DOFs of ITMX calibrated into microns or microrad as per Jamie's ethereal input matrix calculations.
The main result is in the ratio of POS to BUTTER. It tells us that even at nighttime (when this data was taken) we should be able to get some reduction in the arms at 1 Hz.
Whether we can get anything down to 0.1 Hz depends on how the arm control signal compares to the POS signal here. I leave it to Jenne to overlay those traces using a recent Arm lock. |
Attachment 1: null.png
|
|
5138
|
Mon Aug 8 12:08:05 2011 |
rana | Update | SUS | sus at atm |

A plot showing that the daily variation in the OLs is sometimes almost as much as the full scale readout (-1 to +1). |
5174
|
Wed Aug 10 14:35:39 2011 |
rana | Update | PEM | Calibration of Guralp and STS2 |
I'm pretty sure that don't have any ADC's with this gain. It should be +/- 10V for 16 bits. |
5180
|
Wed Aug 10 22:47:22 2011 |
rana | Summary | VAC | Vacuum Workstation (linux3) re-activated |
For some reason the workstation at the vac rack was off and unplugged. Nicole and I plugged its power back in to the EX rack.
I turned it on and it booted up fine; its not dead. To get it on to the network I just made the conversion from 131.215 to 192.168 that Joe had done on all the other computers several months ago.
Now it is showing the Vacuum overview screen correctly again and so Steve no longer has to monopolize one of the Martian laptops over there. |
5312
|
Sat Aug 27 15:47:59 2011 |
rana | Update | CDS | OSEM noise / nullstream and what does it mean for satellites |
In the previous elog of mine, I looked at the nullstream (aka butterfly mode) to find out if the intrinsic OSEM noise is limiting the displacement noise of the interferometer or possibly the Wiener FF performance.
The conclusion was that its not above ~0.2 Hz. Due to the fortuitous breaking of the ITMX magnet, we also have a chance to check the 'bright noise': what the noise is with no magnet to occlude the LED beam.
As expected, the noise spectra with no magnets is less than the calculated nullstream. The attached plot shows the comparison of the LL OSEM (all the bright spectra look basically alike) with the damped
optic spectra from 1 month week ago.
From 0.1 - 10 Hz, the motion is cleanly larger than the noise. Below ~0.2 Hz, its possible that the common mode rejection of the short cavity lengths are ruined by this. We should try to see if the low frequency
noise in the PRC/SRC is explainable with our current knowledge of seismicity and the 2-dimensional 2-poiint correllation functions of the ground.
So, the question is, "Should we try to upgrade the satellite boxes to improve the OSEM sensing noise?" |
Attachment 1: Untitled.png
|
|
5350
|
Tue Sep 6 22:51:53 2011 |
rana | Summary | Cameras | All Camera setups a need upgrading |
I just tried to adjust the ETMY camera and its not very user friendly = NEEDS FIXING.
* Camera view is upside down.
* Camera lens is contacting the lexan viewport cover; this means the focus cannot be adjusted without misaligning the camera.
* There's no strain relief of the camera cables at the can. Needs a rubber cable grommet too.
* There's a BNC "T" in the cable line.
Probably similar issues with some of the other setups; they've had aluminum foil covers for too long. We'll have a camera committee meeting tomorrow to see how to proceed. |
5352
|
Wed Sep 7 00:39:34 2011 |
rana | Update | SUS | ITMX adjustments |
(What we did)
* Moved SUS to edge of table for OSEM adjustment.
* Leveled the table in this temporary tower position.
* Rotated all OSEMs to give some seperation between magnets and LED/PD packages.
* Moved the upper OSEM bracket a little bit upward.
* All the OSEM holding set screws were short with flat heads; this is annoying since we would like to use them more like thumbscrews. Steve took the long set-screws out of the old ITMX cage and we swapped them. Need to order ~100 silver-plated socket head spare/replacements.
* Took pictures of OSEMs.
* Moved tower back to old position.
* Releveled the table (added one rectangular weight in the NW corner of the table).
* Find that ITMX OSEMs were a couple 100 micron out of position; we adjusted them in-situ in the final position of the tower, trying not to rotate them. All mean voltages now are within 100 mV of ideal half-light.
* Back/front EQ positions adjusted by the screw method. bottom/top stops adjusted earlier.
* OSEM cables tied down with copper wire.
* Increased the incident power up to 91 mW going into MC to temporarily make the POX beam more visible.
* The POX beam was checked. It was exiting from the chamber and going through about the center of the viewport. |
5364
|
Wed Sep 7 22:17:04 2011 |
rana | Update | IOO | RF Amp for EOM on PSL Table |
After Steve pointed out the 'deep hoop' issue, we decided to examine putting an RF Amp on the PSL table, between the RF combiner and the triple resonant box.
This will reduce the chances of standing waves in the cables and reduce the radiation induced pick-up in the RF PD and Demod electronics.
We would like to send ~10 dBm from the distribution box to the combiner. We also want to able to get as much as ~33 dBm of drive at 11 and 55 MHz. So the amp should have a gain of ~20-30 dB and an operating range of 10-100 MHz.
Also desirable are low distortion (high IP3) and good reverse isolation ( > 40 dB).
Some possibilities so far (please add your RF Google Results here):
1) Mini-Circuits ZHL-1-2W-S: G = +32 dB, Max Out = +33 dBm, NF = 6 dB, Directivity = 25 dB
2) Mini-Circuits TIA-1000-1R8: G=+40 dB, Max Out = +36 dBm, NF = 15 dB (AC Powered, Inst. Amp), Directivity = 58 dB.
3) Mini-Circuits ZHL-2-8: G = +27dB, Max out = +29 dBm, NF = 6dB, Directivity = 32 dB
4) RFbay MPA-10-40: G = +40dB, Max Out = + 30 dBm, NF = 3.3 dB, Rev Iso = 23 dB
5) No proper stuff from Teledyne Couger
|
5372
|
Fri Sep 9 19:15:17 2011 |
rana | Update | IOO | RF Amp for EOM on PSL Table |
Quote: |
After Steve pointed out the 'deep hoop' issue, we decided to examine putting an RF Amp on the PSL table, between the RF combiner and the triple resonant box.
5) No proper stuff from Teledyne Couger
|
By looking at what Daniel used in the low noise EOM Driver for aLIGO, we found the A2CP2596 from Cougar.
G = +24 dB, NF = 5 dB, Max Out = +37 dBm. It comes in a 2-stage SMA connector package. I've asked Steve to order 2 of them with the appropriate heatsinks. |
5376
|
Sat Sep 10 11:07:37 2011 |
rana | HowTo | SUS | Optical Lever Servo Tuning thoughts |
Now that we are in a moderately stable condition, its time to design the optical lever feedback transfer functions. We should think carefully about how to do this optimally.
In the past, the feedback shape was velocity damping from 0-10 Hz, with some additional resonant gain around the pendulum and stack modes. There were some low pass filters above ~30 Hz. These were all hand tuned.
I propose that we should look into designing optimal feedback loops for the oplevs. In principle, we can do this by defining some optimal feedback cost function and then calculate the poles/zeros in matlab.
How to define the cost function (? please add more notes to this entry):
1) The ERROR signal should be reduced. We need to define a weight function for the ERROR signal: C_1(f) = W_1(f) * (ERR(f)^2)
2) The OL QPDs have a finite sensing noise, so there is no sense in suppressing the signal below this level. Need to determine what the sensing noise is.
3) The feedback signal at high frequencies (30 Hz < f < 300 Hz) should be low passed to prevent adding noise to the interferometer via the A2L coupling. It also doesn't help to reduce this below the level of the seismic noise. The cost function on the feedback should be weighted apprpriately given knowledge about the sensing noise of the OL, the seismic noise (including stack), and the interferometer noise (PRC, SRC, MICH, DARM).
4) The servo should be stable: even if there is a negligible effect on the ERROR signal, we would not want to have more than 10 dB of gain peaking around the UGFs.
5) The OL QPDs are dominated by drift of the stack, laser, etc. at some low frequencies. We should make sure the low frequency feedback is high passed appropriately.
6) Minimize transmitted power rms in single arm lock etc. |
5379
|
Sat Sep 10 18:52:45 2011 |
rana | Update | Computers | conlog getting filled up |
One of the reasons that conlog seems so slow lately is that its been writing 100's of MB of .log files every day since early summer. It looks like the people who have been working on the mdl builds have not been properly adjusting the conlog channel lists. When this is not done conlog just gets filled up with non-control channels like OUT, OUTPUT, OUTMON, etc.
Peter Shawhan has supplied us with many scripts in the conlog directory to clean up these bloated files and fix the channel list. |
5381
|
Sat Sep 10 19:03:57 2011 |
rana | Update | LSC | Y Arm work |
I lined up the Y Arm for locking and then centered the oplevs for ETMY and ITMY.
* The ITMY OL has still got the old style laser. Steve, pleaes swap this one for a HeNe. Also the optical layout seems strange: there are two copies of the laser beam going into the chamber (??). Also, the QPD transimpedance needs to be increase by a factor of ~10. We're only getting ~500 counts per quadrant. Its worth it for someone to re-examine the whole ITMY OL beam layout.
* The ETMY OL beam was coming out but clipping on the mount for the ETMY OL HeNe. This indicates a failure on our part to do the ETMY closeout alignment properly. In fact, I get the feeling from looking around that we overlooked aligning the OL and IPPOS/ANG beams this time. If we're unlucky this could cause us to vent again. I undid part of the laser mount and changed the height on the receiving mirror to get the beam back onto the QPD.
I noticed that there is significant green light now getting into some of the IR PDs; beacuse of this there are weird offsets in the TRY QPD and perhaps elsewhere. We had better purchase some filters to tape over the front of the sensitive IR sensors to prevent the couplling from the green laser.
* There is a beam on IPPOS, but its too big for the detector (this has always been the case). We need to put a 2" lens with a weak focusing power on this path so as to halve the beam size on the detector. Right now its clipping and misleading. There is also a 0.9V offset on the SUM signal. I'm not sure if this readout is working at all.
* I couldn't find any beam on IPANG at all. Not sure what's changed since Kiwamu saw it. |
5383
|
Sat Sep 10 20:30:01 2011 |
rana | Update | IOO | MC trans re-aligned / MC2 shifted mysteriously / MC2 re-aligned |

I re-aligned the beam onto the MC TRANS QPD since Kiwamu had centered the spots on the mirrors. However, I then inspected the MC2F camera. After coming back into the control room I noticed that the MC transmission had gone down by 50% and that the MC2 OSEMs showed a large step. My guess is that somehow the opening and closing of the can shifted the suspension. So I adjusted the MC2 alignment biases to recover the transmitted power (its now ~50000 instead of the ~33000 from Friday). |
5404
|
Wed Sep 14 12:01:05 2011 |
rana | Update | Adaptive Filtering | Modifications to LSC, RFM models, added OAF model |
For the acquisition of the MC_F channel, I suggest taking the FAST_MON BNC output from the blue FSS interface card in the Eurocard crate in the PSL rack. This can then be piped into the 2-pin LEMO plug (Ch. 1) of the Generic Pentek DAQ card which used to acquire the MC_L signal from the MC Servo Board. |
5409
|
Wed Sep 14 20:30:36 2011 |
rana | Update | SUS | Some screens are still bad |
I've found that a few of the screens still have Whited-Out fields due to naming changes (OL SUM and ALS-> TM OFFSET). I attach a screen shot of it.
The OL screens have the wrong SUM names and the IFO ALIGN screen is pointing to the wrong SUS screens.

|
5411
|
Wed Sep 14 22:07:41 2011 |
rana | Update | SUS | ITM Oplevs are broken |
I went to see what was wrong with the ITMs and found that people have been working on them and have left them in a broken state with no elog entry.
This is sad and unacceptable.
Whoever is working on these should post into the elog what the Oplev layout plan is, have someone check it, and ONLY THEN get to work on it.
The layout as it looks tonight is too complicated. With too many optics we will not have a low noise optical lever setup. The new layout should use a bare minimum number of optics and only use very stable mounts.

|
5448
|
Sun Sep 18 14:08:52 2011 |
rana | Update | SUS | Calibration plan for the oplevs |
We don't need a high quality calibration for the optical levers. ~50% accuracy is fine.
For that you can use the OSEM calibration of ~1.7 V/mm (its less than 2 since the OSEMs have been degrading) or you can use the cavity power method that Kakeru used; it worked just fine. There's no benefit in trying for a 1% number for optical levers. |
5466
|
Mon Sep 19 17:45:39 2011 |
rana | Update | SUS | Some screens fixed |
Quote: |
Kiwamu: The bad medm screens have been fixed. There are no blank fields and all the links are correct.
Quote from #5409 |
I've found that a few of the screens still have Whited-Out fields due to naming changes (OL SUM and ALS-> TM OFFSET). I attach a screen shot of it.
The OL screens have the wrong SUM names and the IFO ALIGN screen is pointing to the wrong SUS screens.
|
|
Really? I found this one with ~15 seconds of clicking around.

|