  16242   Fri Jul 9 15:39:08 2021 AnchalSummaryALSSingle Arm Actuation Calibration with IR ALS Beat [Correction]

I did this analysis again by just doing demodulation go 5s time segments of the 60s excitation signal. The major difference is that I was not summing up the sine-cosine multiplied signals, so the error associated was a lot more. If I simply multpy the whole beatnote signal with digital LO created at excitation frequency, divide it up in 12 segments of 5 s each, sum them up individually, then take the mean and standard deviation, I get the answer as:
\frac{6.88 \pm 0.05}{f^2} nm/ctsas opposed to \frac{7.32 \pm 0.03}{f^2} nm/ctsthat was calculated using MICH signal earlier by gautum in 13984.

Attachment 1 shows the scatter plot for the complex calibration factors found for the 12 segments.

My aim in the previous post was however to get a time series of the complex calibration factor from which I can take a noise spectral density measurement of the calibration. I'll still look into how I can do that. I'll have to add a low pass filter to integrate the signal. Then the noise spectrum up to the low pass pole frequency would be available. But what would this noise spectrum really mean? I still have to think a bit about it. I'll put another post soon.

Quote:

We attempted to simulate "oscillator based realtime calibration noise monitoring" in offline analysis with python. This helped us in finding about a factor of sqrt(2) that we were missing earlier in 16171. we measured C1:ALS-BEATX_FINE_PHASE_OUT_HZ_DQ when X-ARM was locked to main laser and Xend green laser was locked to XARM. An excitation signal of amplitude 600 was setn at 619 hz at C1:ITMX_LSC_EXC.

Signal analysis flow:

  • The C1:ALS-BEATX_FINE_PHASE_OUT_HZ_DQ is calibrated to give value of beatntoe frequency in Hz. But we are interested in the fluctuations of this value at the excitation frequency. So the beatnote signal is first high passed with 50 hz cut-off. This value can be reduced a lot more in realtime system. We only took 60s of data and had to remove first 2 seconds for removing transients so we didn't reduce this cut-off further.
  • The I and Q demodulated beatntoe signal is combined to get a complex beatnote signal amplitude at excitation frequency.
  • This signal is divided by cts amplitude of excitation and multiplied by square of excitation frequency to get calibration factor for ITMX in units of nm/cts/Hz^2.
  • The noise spectrum of absolute value of  the calibration factor is plotted in attachment 1, along with its RMS. The calibration factor was detrended linearly so the the DC value was removed before taking the spectrum.
  • So Attachment 1 is the spectrum of noise in calibration factor when measured with this method. The shaded region is 15.865% - 84.135% percentile region around the solid median curves.

We got a value of \frac{7.3 \pm 3.9}{f^2}\, \frac{nm}{cts}.  The calibration factor in use is from \frac{7.32}{f^2} nm/cts from 13984.

Next steps could be to budget this noise while we setup some way of having this calibration factor generated in realitime using oscillators on a FE model. Calibrating actuation of a single optic in a single arm is easy, so this is a good test setup for getting a noise budget of this calibration method.

 

Attachment 1: ITMX_calibration_With_ALS_Beat.pdf
ITMX_calibration_With_ALS_Beat.pdf
  16287   Mon Aug 23 10:17:21 2021 PacoSummaryComputerssystem reboot glitch

[paco]

At 09:34 PST I noted a glitch in the controls room as the machines went down except for c1ioo. Briefly, the video feeds disappeared from the screens, though the screens themselves didn't lose power. At first I though this was some kind of power glitch, but upon checking with Jordan, it most likely was related to some system crash. Coming back to the controls room, I could see the MC reflection beam swinging, but unfortunately all the FE models came down. I noticed that the DAQ status channels were blank.

I ssh into c1ioo no problem and ran "rtcds stop c1ioo c1als c1omc", then "rtcds restart c1x03" to do a soft restart. This worked, but the DAQ status was still blank. I then tried to ssh into c1sus and c1lsc without success, similarly c1iscex and c1iscey were unreachable. I went and did a hard restart on c1iscex by switching it off, then its extension chassis, then unplugging the power cords, then inverting these steps, and could ssh into it from rossa. I ran "rtcds start c1x01" and saw the same blank DAQ status. I noticed the elog was also down... so nodus was also affected?

[paco, anchal]

Anchal got on zoom to offer some assistance. We discovered that the fb1 and nodus were subject to some kind of system reboot at precisely 09:34. The "systemctl --failed" command on fb1 displayed both the daqd_dc.service and rc-local.service as loaded but failed (inactive). Is it a good idea to try and reboot the fb1 machine? ... Anchal was able to bring elog back up from nodus (ergo, this post).

[paco]

Although it probably needs the DAQ service from the fb1 machine to be up and running, I tried running the scripts/cds/rebootC1LSC.sh script. This didn't work. I tried running sudo systemctl restart daqd_dc from the fb1 machine without success. Running systemctl reset-failed "worked" for daqd_dc and rc-local services on fb1 in the sense that they were no longer output from systemctl --failed, but they remained inactive (dead) when running systemctl status on them. Following from  15303   I succeeded in restarting the daqd services. Turned out I needed to manually start the open-mx and mx services in fb1. I rerun the restartC1LSC script without success. The script fails because some machines need to be rebooted by hand.
 

  16303   Mon Aug 30 17:49:43 2021 PacoSummaryLSCXARM POX OLTF

Used diaggui to get OLTF in preparation for optimal system identification / calibration. The excitation was injected at the control point of the XARM loop C1:LSC-XARM_EXC. Attachment 1 shows the TF (red scatter) taken from 35 Hz to 2.3 kHz with 201 points. The swept sine excitation had an envelope amplitude of 50 counts at 35 Hz, 0.2 counts at 100 Hz, and 0.2 at 200 Hz. In purple continous line, the model for the OLTF using all the digital control filters as well as a simple 1 degree of freedom plant (single pole at 0.99 Hz) is overlaid. Note the disagreement of the OLTF "model" at higher frequencies which we may be able to improve upon using vector fitting.

Attachment 2 shows the coherence (part of this initial measurement was to identify an appropriately large frequency range where the coherence is good before we script it).

Attachment 1: XARM_POX_OLTF.pdf
XARM_POX_OLTF.pdf
Attachment 2: XARM_POX_Coh.pdf
XARM_POX_Coh.pdf
  16304   Tue Aug 31 14:55:24 2021 ranaSummaryLSCXARM POX OLTF

this model doesn't seem to include the analog AA, analog AI, digital AA, digital AI, or data transfer delays in the system. I think if you include those you will get more accuracy at high frequencies. Probably Anchal has those included in his DARM loop model?

 

  16306   Wed Sep 1 21:55:14 2021 KojiSummaryGeneralTowards the end upgrade

- Sat amp mod and test: on going (Tega)
- Coil driver mod and test: on going (Tega)

- Acromag: almost ready (Yehonathan)

- IDC10-DB9 cable / D2100641 / IDC10F for ribbon in hand / Dsub9M ribbon brought from Downs / QTY 2 for two ends -> Made 2 (stored in the DSUB connector plastic box)
- IDC40-DB9 cable / D2100640 / IDC40F for ribbon in hand / DB9F solder brought from Downs  / QTY 4 for two ends -> Made 4 0.5m cables (stored in the DSUB connector plastic box)

- DB15-DB9 reducer cable / ETMX2+ETMY2+VERTEX16+NewSOS14 = 34 / to be ordered

- End DAC signal adapter with Dewhitening (with DIFF/SE converter) / to be designed & built
- End ADC adapter (with SE/DIFF converter) / to be designed & built


MISC Ordering

  • 3.5 x Sat Amp Adapter made (order more DSUB25 conns)
    • -> Gave 2 to Tega, 1.5 in the DSUB box
    • 5747842-4 A32100-ND -> ‎5747842-3‎ A32099-ND‎ Qty40
    • 5747846-3 A32125-ND -> ‎747846-3‎ A23311-ND‎ Qty40
  • Tega's sat amp components
    • 499Ω P499BCCT-ND 78 -> Backorder -> ‎RG32P499BCT-ND‎ Qty 100
    • 4.99KΩ TNPW12064K99BEEA 56 -> Qty 100
    • 75Ω YAG5096CT-ND 180 -> Qty 200
    • 1.82KΩ P18391CT-ND 103 -> Qty 120
    • 68 nF P10965-ND 209
  • Order more DB9s for Tega's sat amp adapter 4 units (look at the AA IO BOM) 
    • 4x 8x 5747840-4 DB9M PCB A32092-ND -> 6-747840-9‎ A123182-ND‎ Qty 35
    • 4x 5x 5747844-4 A32117-ND -> Qty 25
    • 4x 5x DB9M ribbon MMR09K-ND -> 8209-8000‎ 8209-8000-ND‎ Qty 25
    • 4x 5x 5746861-4 DB9F ribbon 5746861-4-ND -> 400F0-09-1-00 ‎LFR09H-ND‎ Qty 35
  • Order 18bit DAC AI -> 16bit DAC AI components 4 units
    • 4x 4x 5747150-8 DSUB9F PCB A34072-ND -> ‎D09S24A4PX00LF‎609-6357-ND‎ Qty 20
    • 4x 1x 787082-7 CONN D-TYPE RCPT 68POS R/A SLDR (SCSI Female) A3321-ND -> ‎5787082-7‎ A31814-ND‎ Qty 5
    • 4x 1x 22-23-2021 Connector Header Through Hole 2 position 0.100" (2.54mm)    WM4200-ND -> Qty5

 

 

  16307   Thu Sep 2 17:53:15 2021 PacoSummaryComputerschiara down, vac interlock tripped

[paco, koji, tega, ian]

Today in the morning the name server / network file system running in chiara failed. This resulted in donatella/pianosa/rossa shell prompts to hang forever. It also made sitemap crash and even dropping into a bash shell and just listing files from some directory in the file system froze the computer. Remote ssh sessions on nodus also had the same symptoms.

A little after 1 pm, we started debugging this issue with help from Koji. He suggested we hook a monitor, keyboard, and mouse onto chiara as it should still work locally even if something with the NFS (network file system) failed. We did this and then we tried for a while to unmount the /dev/sdc1/ from /home/cds/ (main file system) and mount /dev/sdb1/ from /media/40mBackup (backup copy) such that they swap places. We had no trouble unmounting the backup drive, but only succeeded in unmounting the main drive with the "lazy" unmount, or running "umount -l". Running "df" we could see that the disk space was 100% used, with only ~ 1 GB of free space which may have been the cause for the issue. After swapping these disks by editing the /etc/fstab file to implement the aforementioned swapping, we rebooted chiara and we recovered the shell prompts in all workstations, sitemap, etc... due to the backup drive mounting. We then started investigating what caused the main drive to fill up that quickly, and noted that weirdly now the capacity was at 85% or about 500GB less than before (after reboot and remount), so some large file was probably dumped into chiara that froze the NFS causing the issue.

At this point we tried opening the PSL shutter to recover the IMC. The shutter would not open and we suspected the vacuum interlock was still tripped... and indeed there was an uncleared error in the VAC screen. So with Koji's guidance we walked to the c1vac near the HV station and did the following at ~ 5:13 PM -->

  1. Open V4; apart from a brief pressure spike in PTP2, everything looked ok so we proceeded to
  2. Open V1; P2 spiked briefly and then started to drop. Then, Koji suggested that we could
  3. Close V4; but we saw P2 increasing by a factor of~ 10 in a few seconds, so we
  4. Reopened V4;

We made sure that P1a (main vacuum pressure) was dropping and before continuing we decided to look back to see what the nominal vacuum state was that we should try to restore.

We are currently searching the two systems for diffrences to see if we can narrow down the culprit of the failure.

 

  16312   Thu Sep 2 21:21:14 2021 KojiSummaryComputersVacuum recovery 2

Attachment 1:
We are pumping the main volume with TP2. Once P1a reached the pressure ~2.2mtorr, we could open the PSL shutter. The TP2 voltage went up once but came down to ~20V. It's close to nominal now.
We wondered if we should use TP3 or not. I checked the vacuum pressure trends and found that the annulus pressures were going up. So we decided to open the annulus valves.

Attachment 2:
The current vacuum status is as shown in the MEDM screenshot.

There is no trend data of the valve status (sad)

Attachment 1: Screenshot_2021-09-02_21-20-24.png
Screenshot_2021-09-02_21-20-24.png
Attachment 2: Screenshot_2021-09-02_21-20-48.png
Screenshot_2021-09-02_21-20-48.png
  16313   Thu Sep 2 21:49:03 2021 PacoSummaryComputerschiara down, vac interlock tripped

[tega, paco]

We found the files that took excess space in the chiara filesystem (see Attachment 1). They were error files from the summary pages that were ~ 50 GB in size or so located under /home/cds/caltech/users/public_html/detcharsummary/logs/. We manually removed them and then copied the rest of the summary page contents into the main file system drive (this is to preserve the information backup before it gets deleted by the cron job at the end of today) and checked carefully to identify the actual issue for why these files were as large in the first place.

We then copied the /detcharsummary directory from /media/40mBackup into /home/cds to match the two disks.

Attachment 1: 2021-09-02_21-51-15.png
2021-09-02_21-51-15.png
  16314   Fri Sep 3 02:03:15 2021 TegaSummaryComputersStrip down large error files

Also deleted the ~50GB error files from ldas to prevent rsync from copying them to nodus again. With the new update to GWsumm, there are new error messages that initially didn't seem to affect the summary pages functionality, but in the extreme case can populated the error files the repeated warnings on the form "Loading: FrSerData", "Loading: FrSerData::n4294967295", "Loading: FrSummary","Loading: FrSerDataLoading: FrSerData" and many more combinations until we get file sizes of the order of ~50GB. So I have updated the checkstatus script to parse the error files and strip out the majority of these error messages. Work is ongoing to get them all.

In light of these large files generation, I decided to look in the summary pages folder to see if there are other large files that we need to keep track of and it turns there are indeed a collection of files in the archive folder that bloats the summary pages on ldas to ~1TB. Luckily these are not synced to nodus so no problem here. However, since the beginning of the year, the archive folders that hold data used for each day's computation have not been cleared. We have a script for doing this but it has not been run for a while now and it only delete archive files for a specific month which is hardcoded to two months from the date the file is run. I have modified the code to allow archive deletion for a range of months so we can clear data from Jan to July. 

Quote:

[tega, paco]

We found the files that took excess space in the chiara filesystem (see Attachment 1). They were error files from the summary pages that were ~ 50 GB in size or so located under /home/cds/caltech/users/public_html/detcharsummary/logs/. We manually removed them and then copied the rest of the summary page contents into the main file system drive (this is to preserve the information backup before it gets deleted by the cron job at the end of today) and checked carefully to identify the actual issue for why these files were as large in the first place.

We then copied the /detcharsummary directory from /media/40mBackup into /home/cds to match the two disks.

 

  16315   Tue Sep 7 18:00:54 2021 TegaSummaryCalibrationSystem Identification via line injection

[paco]

This morning, I spent some time restoring the jupyter notebook server running in allegra. This server was first set up by Anchal to be able to use the latest nds python API tools which is handy for the calibration stuff. The process to restore the environment was to run "source ~/bashrc.d/*" to restore some of the aliases, variables, paths, etc... that made the nds server work. I then ran ssh -N -f -L localhost:8888:localhost:8888 controls@allegra from pianosa and carry on with the experiment.


[paco, hang, tega]

We started a notebook under /users/paco/20210906_XARM_Cal/XARM_Cal.ipynb on which the first part was doing the following;

  • Set up list of excitations for C1:LSC-XARM_EXC (for example three sine waveforms) using awg.py
  • Make sure the arm is locked
  • Read a reference time trace of the C1:LSC-XARM_IN2 channel for some duration
  • Start excitations (one by one at the moment, ramptime ~ 3 seconds, same duration as above)
  • Get data for C1:LSC-XARM_IN2 for an equal duration (raw data in Attachment #1)
  • Generate the excitation sine and cosine waveforms using numpy and demodulate the raw timeseries using a 4th order lowpass filter with fc ~ 10 Hz
  • Estimate the correct demod phase by computing arctan(Q / I) and rerunning the demodulation to dump the information into the I quadrature (Attachment #2).
  • Plot the estimated ASD of all the quadratures (Attachment #3)

[paco, hang, tega]

Estimation of open loop gain:

  • Grab data from the C1:LSC-XARM_IN1 and C1:LSC-XARM_IN2 test points
  • Infer excitation from their differnce, i.e. C1:LSC-XARM_EXC = C1:LSC-XARM_IN2 - C1:LSC-XARM_IN1
  • Compute the open loop gain as follows : G(f) = csd(EXC,IN1)/csd(EXC,IN2), where csd computes the cross spectra density of the input arguments
  • For the uncertainty in G, dG, we repeat steps (1) to (3) with & without signal injection in the C1:LSC-XARM_EXC channel. In the absence of signal injection, the signal in C1:LSC-XARM_IN2 is of the form: Y_ref = Noise/(1-G), whereas with nonzero signal injection, the signal in C1:LSC-XARM_IN2 has the form: Y_cal = EXC/(1-G) + Noise/(1-G), so their ratio, Y_cal/Y_ref = EXC/Noise, gives the SNR, which we can then invert to give the uncertainty in our estimation of G, i.e dG = Y_ref/Y_cal.
  • For the excitation at 53 Hz, our measurtement for the open loop gain comes out to about 5 dB whiich is consistent with previous measurement.
  • We seem to have an SNR in excess of 100 at measurement time of 35 seconds and 1 count of amplitude which gives a relative uncertainty of G of 0.1%
  • The analysis details are ongoing. Feedback is welcome.
Attachment 1: raw_timeseries.pdf
raw_timeseries.pdf
Attachment 2: demod_signals.pdf
demod_signals.pdf
Attachment 3: cal_noise_asd.pdf
cal_noise_asd.pdf
  16318   Thu Sep 9 09:54:41 2021 StephenSummaryBHDBHD OMC invacuum wiring - cable lengths

[Koji, Stephen - updated 30 September]

Cable lengths task - in vacuum cabling for the green section (new, custom for 40m) and yellow section (per aLIGO, except likely with cheaper FEP ribbon cable material) from 40m/16198. These arethe myriad of cables extending from the in vacuum flange to the aLIGO-style on-table Cable Stand (think, for example, D1001347), then from the cable stand to the OMCs.

a) select a position for the cable stand.

 - Koji and I discussed and elected to place in the (-X, -Y) corner of the table (Northwest in the typical diagram) and near the table edge. This is adjacent to the intended exit flange for the last cable.

b) measure distances and cable routing approximations for cable bracket junctions

- Near OMC bracket to the cable stand, point to point = 17.2, routing estimate = 24.4.
- Far OMC bracket to the cable stand, point to point = 20.5, routing estimate = 32.2.

  - Recommendation = 48" for all green section cables (using one length for each OMC, with extra slack to account for routing variation).

c) (outdated - see item (b) and attachment 3) measure distances (point to point) and cable routing approximations for all items.

 +X OMC (long edge aligned with +Y beam axis) (overview image in Attachment 1)

- QPDs to the cable stand, point to point = 12, routing estimate = 20.
- DCPDs to the cable stand, point to point = 25, routing estimate = 32.
- PZTs to the cable stand, point to point = 21, routing estimate = 32.

+Y OMC (long edge aligned with +Y beam axis) (overview image in Attachment 1)

- QPDs to the cable stand, point to point = 16, routing estimate = 23.
- DCPDs to the cable stand, point to point = 26, routing estimate = 38.
- PZTs to the cable stand, point to point = 24, routing estimate = 33.

Cable stand to flange (Attachment 2) (specific image in Attachment 2)

- point to point = 35, routing estimate = 42

  - Recommendation = 120" for all yellow section cables, per Koji's preferences for zigzag cable routing on stack and coiling of slack.

Attachment 1: bhd_cable_length_check_cable_bracket_to_components.png
bhd_cable_length_check_cable_bracket_to_components.png
Attachment 2: bhd_cable_length_check_flange_to_cable_bracket.png
bhd_cable_length_check_flange_to_cable_bracket.png
Attachment 3: bhd_cable_length_check_cable_bracket_to_omc_bracket.png
bhd_cable_length_check_cable_bracket_to_omc_bracket.png
  16323   Mon Sep 13 17:05:04 2021 TegaSummaryPEMInfrasensing temperature sensor modbus configuration

Anchal mentioned it would be good to put more details about how I arrived at the values needed to configure the modbus drive for the temperature sensor, since this information is not in the manual and is hard to find on the internet, so here is a breakdown.

So the generic format is:

drvAsynIPPortConfigure("<TCP_PORT_NAME>","<UNIT_IP_ADDRESS>:502",0,0,1)
modbusInterposeConfig("<TCP_PORT_NAME>",0,5000,0)
drvModbusAsynConfigure("<PORT_NAME>","<TCP_PORT_NAME>",<slaveAddress>,<modbusFunction>,<modbusStartAddress>,<modbusLength>,<dataType>,<pollMsec>,<plcType>)

which in our case become:

drvAsynIPPortConfigure("c1pemxendtemp","192.168.113.240:502",0,0,1)
modbusInterposeConfig("c1pemxendtemp",0,5000,0)
drvModbusAsynConfigure("C1PEMXENDTEMP","c1pemxendtemp",0,4,199,2,0,1000,"ServerCheck")

As can be seen, the parameters of the first two functions "drvAsynIPPortConfigure" and "modbusInterposeConfig" are straight forward, so we restrict our discussion to the case of third function "drvModbusAsynConfigure". Well, after hours of trolling the internet, I was able to piece together a coherent picture of what needs doing and I have summarised them in the table below.

 

drvModbusAsynConfigure

Once the asyn IP or serial port driver has been created, and the modbusInterpose driver has been configured, a modbus port driver is created with the following command:

drvModbusAsynConfigure(portName,                # used by channel definitions in .db file to reference this unit)
                       tcpPortName,             # reference to portName created with drvAsynIPPortConfigure command
                       slaveAddress,            # 
                       modbusFunction,          # 
                       modbusStartAddress,      # 
                       modbusLength,            # length in dataType units
                       dataType,                # 
                       pollMsec,                # how frequently to request a value in [ms]
                       plcType);                #

drvModbusAsynConfigure command
Parameter Data type Description
portName string Name of the modbus port to be created.
 
tcpPortName string Name of the asyn IP or serial port previously created.

tcpPortName = { 192.168.113.240:502192.168.113.241:502192.168.113.242:502 }
 
slaveAddress int The address of the Modbus slave. This must match the configuration of the Modbus slave (PLC) for RTU and ASCII. For TCP the slave address is used for the "unit identifier", the last field in the MBAP header. The "unit identifier" is ignored by most PLCs, but may be required by some.

ServersCheck API ignores this value, as confirmed with pymodbus query, so set to default value: 
slaveAddress = 0
 
modbusFunction int

modbus supports the following 8 Modbus function codes:

Modbus Function Codes
Access Function description Function code
Bit access Read Coils 1
Bit access Read Discrete Inputs 2
Bit access Write Single Coil 5
Bit access Write Multiple Coils 15
16-bit word access Read Input Registers 4
16-bit word access Read Holding Registers 3
16-bit word access Write Single Register 6
16-bit word access Write Multiple Registers 16
modbusStartAddress int Start address for the Modbus data segment to be accessed.
(0-65535 decimal, 0-0177777 octal).

Modbus addresses are specified by a 16-bit integer address. The location of inputs and outputs within the 16-bit address space is not defined by the Modbus protocol, it is vendor-specific. Note that 16-bit Modbus addresses are commonly specified with an offset of 400001 (or 300001). This offset is not used by the modbus driver, it uses only the 16-bit address, not the offset.

For ServersCheck, the offset is "30001", so that

modbusStartAddress = 30200 - 30001 = 199

modbusLength int The length of the Modbus data segment to be accessed.
This is specified in bits for Modbus functions 1, 2, 5 and 15.
It is specified in 16-bit words for Modbus functions 3, 4, 6 and 16.
Length limit is 2000 for functions 1 and 2, 1968 for functions 5 and 15,
125 for functions 3 and 4, and 123 for functions 6 and 16.

ServersCheck uses two's complement 32-bits word (with big-endian byte order & little-endian word order) format to store floating-point data, as confirmed with pymodbus query, so that:

modbusLength = 2
 
modbusDataType int The modbusDataType is used to tell the driver the format of the Modbus data. The driver uses this information to convert the number between EPICS and Modbus. Data is transferred to and from EPICS as epicsUInt32, epicsInt32, and epicsFloat64 numbers.

Modbus data type:
0 = binary, twos-complement format
1 = binary, sign and magnitude format
2 = BCD, unsigned
3 = BCD, signed

Some Modbus devices (including ServersCheck) use floating point numbers, typically by storing a 32-bit float in two consecutive 16-bit registers. This is not supported by the Modbus specification, which only supports 16-bit registers and single-bit data. The modbus driver does not directly support reading such values, because the word order and floating point format is not specified.

Note that if it is desired to transmit BCD numbers untranslated to EPICS over the asynInt32 interface, then data type 0 should be used, because no translation is done in this case. 

For ServersCheck, we wish to transmit the untranslated data, so:

modbusDataType = 0
 
pollMsec int Polling delay time in msec for the polling thread for read functions.
For write functions, a non-zero value means that the Modbus data should
be read once when the port driver is first created.

ServersCheck recommends setting sensor polling interval between 1-5 seconds, so we can try:

pollMsec = 1000
 
plcType string Type of PLC (e.g. Koyo, Modicon, etc.).
This parameter is currently used only to print information in asynReport.
In the future it could be used to modify the driver behavior for a specific PLC.

plcType = "ServersCheck"
 

 

Useful links

https://nodus.ligo.caltech.edu:8081/40m/16214

https://nodus.ligo.caltech.edu:8081/40m/16269

https://nodus.ligo.caltech.edu:8081/40m/16270

https://nodus.ligo.caltech.edu:8081/40m/16274

 

http://manuals.serverscheck.com/InfraSensing_Sensors_Platform.pdf

http://manuals.serverscheck.com/InfraSensing_Modbus_manualv5.pdf

https://community.serverscheck.com/discussion/comment/7419#Comment_7419

 

https://wiki-40m.ligo.caltech.edu/CDS/SlowControls

https://www.slac.stanford.edu/grp/ssrl/spear/epics/site/modbus/modbusDoc.html#Creating_a_modbus_port_driver

 

https://github.com/riptideio/pymodbus

 

https://en.wikipedia.org/wiki/Modbus

https://deltamotion.com/support/webhelp/rmctools/Communications/Ethernet/Supported_Protocols/Ethernet_Modbus_TCP.htm

  16329   Tue Sep 14 17:19:38 2021 PacoSummaryPEMExcess seismic noise in 0.1 - 0.3 Hz band

For the past couple of days the 0.1 to 0.3 Hz RMS seismic noise along BS-X has increased. Attachment 1 shows the hour trend in the last ~ 10 days. We'll keep monitoring it, but one thing to note is how uncorrelated it seems to be from other frequency bands. The vertical axis in the plot is in um / s

Attachment 1: SEIS_2021-09-14_17-33-12.png
SEIS_2021-09-14_17-33-12.png
  16331   Tue Sep 14 19:12:03 2021 KojiSummaryPEMExcess seismic noise in 0.1 - 0.3 Hz band

Looks like this increase is correlated for BS/EX/EY. So it is likely to be real.

Comparison between 9/15 (UTC) (Attachment 1) and 9/10 (UTC) (Attachment 2)

Attachment 1: C1-ALL_393F21_SPECTRUM-1315699218-86400.png
C1-ALL_393F21_SPECTRUM-1315699218-86400.png
Attachment 2: C1-ALL_393F21_SPECTRUM-1315267218-86400.png
C1-ALL_393F21_SPECTRUM-1315267218-86400.png
  16334   Wed Sep 15 23:53:54 2021 KojiSummaryGeneralTowards the end upgrade

Ordered compoenents are in.

- Made 36 more Sat Amp internal boards (Attachment 1). Now we can install the adapters to all the 19 sat amp units.

- Gave Tega the components for the sat amp adapter units. (Attachment 2)

- Gave Tega the componennts for the sat amp / coil driver modifications.

- Made 5 PCBs for the 16bit DAC AI rear panel interface (Attachment 3)

Attachment 1: P_20210915_231308.jpg
P_20210915_231308.jpg
Attachment 2: P_20210915_225039.jpg
P_20210915_225039.jpg
Attachment 3: P_20210915_224341.jpg
P_20210915_224341.jpg
  16343   Mon Sep 20 12:20:31 2021 PacoSummarySUSPRM and BS Angular Actuation transfer function magnitude measurements

[yehonathan, paco, anchal]

We attempted to find any symptoms for actuation problems in the PRMI configuration when actuated through BS and PRM.

Our logic was to check angular (PIT and YAW) actuation transfer function in the 30 to 200 Hz range by injecting appropriately (f^2) enveloped excitations in the SUS-ASC EXC points and reading back using the SUS_OL (oplev) channels.

From the controls, we first restored the PRMI Carrier to bring the PRM and BS to their nominal alignment, then disabled the LSC output (we don't need PRMI to be locked), and then turned off the damping from the oplev control loops to avoid supressing the excitations.

We used diaggui to measure the 4 transfer functions magnitudes PRM_PIT, PRM_YAW, BS_PIT, BS_YAW, as shown below in Attachments #1 through #4. We used the Oplev calibrations to plot the magnitude of the TFs in units of urad / counts, and verified the nominal 1/f^2 scaling for all of them. The coherence was made as close to 1 as possible by adjusting the amplitude to 1000 counts, and is also shown below. A dip at 120 Hz is probably due to line noise. We are also assuming that the oplev QPDs have a relatively flat response over the frequency range below.

Attachment 1: PRM_PIT_ACT_TF.pdf
PRM_PIT_ACT_TF.pdf
Attachment 2: PRM_YAW_ACT_TF.pdf
PRM_YAW_ACT_TF.pdf
Attachment 3: BS_PIT_ACT_TF.pdf
BS_PIT_ACT_TF.pdf
Attachment 4: BS_YAW_ACT_TF.pdf
BS_YAW_ACT_TF.pdf
  16345   Mon Sep 20 14:22:00 2021 ranaSummarySUSPRM and BS Angular Actuation transfer function magnitude measurements

I suggest plotting all the traces in the plot so we can see their differences. Also remove the 1/f^2 slope so that we can see small differences. Since the optlev servos all have low pass filters around 15-20 Hz, its not necessary to turn off the optlev servos for this measurement.

I think that based on the coherence and the number of averages, you should also be able to use Bendat and Piersol so estimate the uncertainy as a function of frequency. And we want to see the comparison coil-by-coil, not in the DoF basis.

4 sweeps for BS and 4 sweeps for PRM.

  16348   Mon Sep 20 15:42:44 2021 Ian MacMillanSummaryComputersQuantization Code Summary

This post serves as a summary and description of code to run to test the impacts of quantization noise on a state-space implementation of the suspension model.

Purpose: We want to use a state-space model in our suspension plant code. Before we can do this we want to test to see if the state-space model is prone to problems with quantization noise. We will compare two models one for a standard direct-ii filter and one with a state-space model and then compare the noise from both. 

Signal Generation:

First I built a basic signal generator that can produce a sine wave for a specified amount of time then can produce a zero signal for a specified amount of time. This will let the model ring up with the sine wave then decay away with the zero signal. This input signal is generated at a sample rate of 2^16 samples per second then stored in a numpy array. I later feed this back into both models and record their results.

State-space Model:

The code can be seen here

The state-space model takes in the list of excitation values and feeds them through a loop that calculates the next value in the output.

Given that the state-space model follows the form

  \dot{x}(t)=\textbf{A}x(t)+ \textbf{B}u(t)   and  y(t)=\textbf{C}x(t)+ \textbf{D}u(t) ,

the model has three parts the first equation, an integration, and the second equation.

  1. The first equation takes the input x and the excitation u and generates the x dot vector shown on the left-hand side of the first state-space equation.
  2. The second part must integrate x to obtain the x that is found in the next equation. This uses the velocity and acceleration to integrate to the next x that will be plugged into the second equation
  3. The second equation in the state space representation takes the x vector we just calculated and then multiplies it with the sensing matrix C. we don't have a D matrix so this gives us the next output in our system

This system is the coded form of the block diagram of the state space representation shown in attachment 1

Direct-II Model:

The direct form 2 filter works in a much simpler way. because it involves no integration and follows the block diagram shown in Attachment 2, we can use a single difference equation to find the next output. However, the only complication that comes into play is that we also have to keep track of the w(n) seen in the middle of the block diagram. We use these two equations to calculate the output value

y[n]=b_0 \omega [n]+b_1 \omega [n-1] +b_2 \omega [n-2],  where w[n] is  \omega[n]=x[n] - a_1 \omega [n-1] -a_2 \omega[n-2]

Bit length Control:

To control the bit length of each of the models I typecast all the inputs using np.float then the bit length that I want. This simulates the computer using only a specified bit length. I have to go through the code and force the code to use 128 bit by default. Currently, the default is 64 bit which so at the moment I am limited to 64 bit for highest bit length. I also need to go in and examine how numpy truncates floats to make sure it isn't doing anything unexpected.

Bode Plot: 

The bode plot at the bottom shows the transfer function for both the IIR model and the state-space model. I generated about 100 seconds of white noise then computed the transfer function as 

G(f) = \frac{P_{csd}(f)}{P_{psd}(f)}

which is the cross-spectral density divided by the power spectral density. We can see that they match pretty closely at 64 bits. The IIR direct II model seems to have more noise on the surface but we are going to examine that in the next elog

 

Attachment 1: 472px-Typical_State_Space_model.svg.png
472px-Typical_State_Space_model.svg.png
Attachment 2: Biquad_filter_DF-IIx.svg.png
Biquad_filter_DF-IIx.svg.png
Attachment 3: SS-IIR-TF.pdf
SS-IIR-TF.pdf
  16351   Tue Sep 21 11:09:34 2021 AnchalSummaryCDSXARM YARM UGF Servo and Oscillators added

I've updated the c1LSC simulink model to add the so-called UGF servos in the XARM and YARM single arm loops as well. These were earlier present in DARM, CARM, MICH and PRCL loops only. The UGF servo themselves serves a larger purpose but we won't be using that. What we have access to now is to add an oscillator in the single arm and get realtime demodulated signal before and after the addition of the oscillator. This would allow us to get the open loop transfer function and its uncertaintiy at particular frequencies (set by the oscillator) and would allow us to create a noise budget on the calibration error of these transfer functions.

 

The new model has been committed locally in the 40m/RTCDSmodels git repo. I do not have rights to push to the remote in git.ligo. The model builds, installs and starts correctly.

  16352   Tue Sep 21 11:13:01 2021 PacoSummaryCalibrationXARM calibration noise

Here are some plots from analyzing the C1:LSC-XARM calibration. The experiment is done with the XARM (POX) locked, a single line is injected at C1:LSC-XARM_EXC at f0 with some amplitude determined empirically using diaggui and awggui tools. For the analysis detailed in this post, f0 = 19 Hz, amp = 1 count, and gain = 300 (anything larger in amplitude would break the lock, and anything lower in frequency would not show up because of loop supression). Clearly, from Attachment #3 below, the calibration line can be detected with SNR > 1.

We read the test point right after the excitation C1:LSC-XARM_IN2 which, in a simplified loop will carry the excitation suppressed by 1 - OLTF, the open loop transfer function. The line is on for 5 minutes, and then we read for another 5 minutes but with the excitation off to have a reference. Both the calibration and reference signal time series are shown in Attachment #1 (decimated by 8). The corresponding ASDs are shown in Attachment #2. Then, we demodulate at 19 Hz and a 30 Hz, 4th-order butterworth LPF, and get an I and Q timeseries (shown in Attachment #3). Even though they look similar, the Q is centered about 0.2 counts, while the I is centered about 0.0. From this time series, we can of course show the noise ASDs in Attachment #3.


The ASD uncertainty bands in the last plot are statistical estimates and depend on the number of segments used in estimating the PSD. A thing to note is that the noise features surrounding the signal ASD around f0 are translated into the ASD in the demodulated signals, but now around dc. I guess from Attachment #3 there is no difference in the noise spectra around the calibration line with and without the excitation. This is what I would have expected from a linear system. If there was a systematic contribution, I would expect it to show at very low frequencies.

Attachment 1: XARM_signal_asd.pdf
XARM_signal_asd.pdf
Attachment 2: XARM_demod_timeseries.pdf
XARM_demod_timeseries.pdf
Attachment 3: XARM_demod_asds.pdf
XARM_demod_asds.pdf
Attachment 4: XARM_cal_0921_timeseries.pdf
XARM_cal_0921_timeseries.pdf
  16353   Wed Sep 22 11:43:04 2021 ranaSummaryCalibrationXARM calibration noise

I would expect to see some lower frequency effects. i.e. we should look at the timeseries of the demod with the excitation on and off.

I would guess tat the exc on should show us the variations in the optical gain below 3 Hz, whereas the exc off would not show it.

Maybe you should do some low pass filtering on the time series you have to see the ~DC effects? Also, reconsider your AA filter design: how do you quantitatively choose the cutoff frequency and stopband depth?

  16354   Wed Sep 22 12:40:04 2021 AnchalSummaryCDSXARM YARM UGF Servo and Oscillators shifted to OAF

To reduce burden on c1lsc, I've shifted the added UGF block to to c1oaf model. c1lsc had to be modified to allow addition of an oscillator in the XARm and YARM control loops and take out test points before and after the addition to c1oaf through shared memory IPC to do realtime demodulation in c1oaf model.

The new models built and installed successfully and I've been able to recover both single arm locks after restarting the computers.

 

  16355   Wed Sep 22 14:22:35 2021 Ian MacMillanSummaryComputersQuantization Noise Calculation Summary

Now that we have a model of how the SS and IIR filters work we can get to the problem of how to measure the quantization noise in each of the systems. Den Martynov's thesis talks a little about this. from my understanding: He measured quantization noise by having two filters using two types of variables with different numbers of bits. He had one filter with many more bits than the second one. He fed the same input signal to both filters then recorded their outputs x_1 and x_2, where x_2 had the higher number of bits. He then took the difference x_1-x_2. Since the CDS system uses double format, he assumes that quantization noise scales with mantissa length. He can therefore extrapolate the quantization noise for any mantissa length.

Here is the Code that follows the following procedure (as of today at least)

This problem is a little harder than I had originally thought. I took Rana's advice and asked Aaron about how he had tackled a similar problem. We came up with a procedure explained below (though any mistakes are my own):

  1. Feed different white noise data into three of the same filter this should yield the following equation: \textbf{S}_i^2 =\textbf{S}_{ni}^2+\textbf{S}_x^2, where \textbf{S}_i^2 is the power spectrum of the output for the ith filter,  \textbf{S}_{ni}^2 is the noise filtered through an "ideal" filter with no quantization noise, and  \textbf{S}_x^2 is the power spectrum of the quantization noise. Since we are feeding random noise into the input the power of the quantization noise should be the same for all three of our runs.
  2. Next, we have our three outputs:  \textbf{S}_1^2,  \textbf{S}_2^2, and  \textbf{S}_3^2 that follow the equations: 

\textbf{S}_1^2 =\textbf{S}_{n1}^2+\textbf{S}_x^2

\textbf{S}_2^2 =\textbf{S}_{n2}^2+\textbf{S}_x^2

\textbf{S}_3^2 =\textbf{S}_{n3}^2+\textbf{S}_x^2

From these three equations, we calculate the three quantities: \textbf{S}_{12}^2\textbf{S}_{23}^2, and \textbf{S}_{13}^2 which are calculated by:

\textbf{S}_{12}^2 =\textbf{S}_{1}^2-\textbf{S}_2^2\approx \textbf{S}_{n1}^2 -\textbf{S}_{n2}^2

\textbf{S}_{23}^2 =\textbf{S}_{2}^2-\textbf{S}_3^2\approx \textbf{S}_{n2}^2 -\textbf{S}_{n3}^2

\textbf{S}_{13}^2 =\textbf{S}_{1}^2-\textbf{S}_3^2\approx \textbf{S}_{n1}^2 -\textbf{S}_{n3}^2

from these quantities, we can calculate three values: \bar{\textbf{S}}_{n1}^2\bar{\textbf{S}}_{n2}^2, and \bar{\textbf{S}}_{n3}^2 since these are just estimates we are using a bar on top. These are calculated using:

\bar{\textbf{S}}_{n1}^2\approx\frac{1}{2}(\textbf{S}_{12}^2+\textbf{S}_{13}^2+\textbf{S}_{23}^2)

\bar{\textbf{S}}_{n2}^2\approx\frac{1}{2}(-\textbf{S}_{12}^2+\textbf{S}_{13}^2+\textbf{S}_{23}^2)

\bar{\textbf{S}}_{n3}^2\approx\frac{1}{2}(\textbf{S}_{12}^2+\textbf{S}_{13}^2-\textbf{S}_{23}^2)

using these estimates we can then estimate  \textbf{S}_{x}^2  using the formula:

\textbf{S}_{x}^2 \approx \textbf{S}_{1}^2 - \bar{\textbf{S}}_{n1}^2 \approx \textbf{S}_{2}^2 - \bar{\textbf{S}}_{2}^2 \approx \textbf{S}_{3}^2 - \bar{\textbf{S}}_{n3}^2

we can average the three estimates for  \textbf{S}_{x}^2  to come up with one estimate.

This procedure should be able to give us a good estimate of the quantization noise. However, in the graph shown in the attachments below show that the noise follows the transfer function of the model to begin with. I would not expect this to be true so I believe that there is an error in the above procedure or in my code that I am working on finding. I may have to rework this three-corner hat approach. I may have a mistake in my code that I will have to go through.

I would expect the quantization noise to be flatter and not follow the shape of the transfer function of the model. Instead, we have what looks like just the result of random noise being filtered through the model. 

Next steps:

The first real step is being able to quantify the quantization noise but after I fix the issues in my code I will be able to start liking at optimal model design for both the state-space model and the direct form II model. I have been looking through the book "Quantization noise" by Bernard Widrow and Istvan Kollar which offers some good insights on how to minimize quantization noise. 

Attachment 1: IIR64-bitnoisespectrum.pdf
IIR64-bitnoisespectrum.pdf
  16358   Thu Sep 23 15:29:11 2021 PacoSummarySUSPRM and BS Angular Actuation transfer function magnitude measurements

[Anchal, Paco]

We had a second go at this with an increased number of averages (from 10 to 100) and higher excitation amplitudes (from 1000 to 10000). We did this to try to reduce the relative uncertainty a-la-Bendat-and-Pearsol

\delta G / G = \frac{1}{\gamma \sqrt{n_{\rm avg}}}

where \gamma, n_{\rm avg} are the coherence and number of averages respectively. Before, this estimate had given us a ~30% relative uncertainty and now it has been improved to ~ 10%. The re-measured TFs are in Attachment #1. We did 4 sweeps for each optic (BS, PRM) and removed the 1/f^2 slope for clarity. We note a factor of ~ 4 difference in the magnitude of the coil to angle TFs from BS to PRM (the actuation strength in BS is smaller).


For future reference:

With complex G, we get complex error in G using the formula above. To get uncertainity in magnitude and phase from real-imaginary uncertainties, we do following (assuming the noise in real and imaginary parts of the measured transfer function are incoherent with each other):
G = \alpha + i\beta

\delta G = \delta\alpha + i\delta \beta

\delta |G| = \frac{1}{|G|}\sqrt{\alpha^2 \delta\alpha^2 + \beta^2 \delta \beta^2}

\delta(\angle G) = \frac{1}{|G|^2}\sqrt{\alpha^2 \delta\alpha^2 + \beta^2 \delta\beta^2} = \frac{\delta |G|}{|G|}

Attachment 1: BS_PRM_ANG_ACT_TF.pdf
BS_PRM_ANG_ACT_TF.pdf BS_PRM_ANG_ACT_TF.pdf BS_PRM_ANG_ACT_TF.pdf BS_PRM_ANG_ACT_TF.pdf
  16360   Mon Sep 27 12:12:15 2021 Ian MacMillanSummaryComputersQuantization Noise Calculation Summary

I have not been able to figure out a way to make the system that Aaron and I talked about. I'm not even sure it is possible to pull the information out of the information I have in this way. Even the book uses a comparison to a high precision filter as a way to calculate the quantization noise:

"Quantization noise in digital filters can be studied in simulation by comparing the behavior of the actual quantized digital filter with that of a refrence digital filter having the same structure but whose numerical calculations are done extremely accurately."
-Quantization Noise by Bernard Widrow and Istvan Kollar (pg. 416)

Thus I will use a technique closer to that used in Den Martynov's thesis (see appendix B starting on page 171). A summary of my understanding of his method is given here:

A filter is given raw unfiltered gaussian data f(t) then it is filtered and the result is the filtered data x(t) thus we get the result: f(t)\rightarrow x(t)=x_N(t)+x_q(t)  where x_N(t) is the raw noise filtered through an ideal filter and x_q(t) is the difference which in this case is the quantization noise. Thus I will input about 100-1000 seconds of the same white noise into a 32-bit and a 64-bit filter. (hopefully, I can increase the more precise one to 128 bit in the future) then I record their outputs and subtract the from each other. this should give us the Quantization error e(t):
e(t)=x_{32}(t)-x_{64}(t)=x_{N_{32}}(t)+x_{q_{32}}(t) - x_{N_{64}}(t)-x_{q_{64}}(t)
and since x_{N_{32}}(t)=x_{N_{64}}(t) because they are both running through ideal filters:
e(t)=x_{N}(t)+x_{q_{32}}(t) - x_{N}(t)-x_{q_{64}}(t)
e(t)=x_{q_{32}}(t) -x_{q_{64}}(t)
and since in this case, we are assuming that the higher bit-rate process is essentially noiseless we get the Quantization noise x_{q_{32}}(t).

If we make some assumptions, then we can actually calculate a more precise version of the quantization noise:

"Since aLIGO CDS system uses double precision format, quantization noise is extrapolated assuming that it scales with mantissa length"
-Denis Martynov's Thesis (pg. 173)

From this assumption, we can say that the noise difference between the 32-bit and 64-bit filter outputs:  x_{q_{32}}(t)-x_{q_{64}}(t)  is proportional to the difference between their mantissa length. by averaging over many different bit lengths, we can estimate a better quantization noise number.

I am building the code to do this in this file

  16361   Mon Sep 27 16:03:15 2021 Ian MacMillanSummaryComputersQuantization Noise Calculation Summary

I have coded up the procedure in the previous post: The result does not look like what I would expect. 

As shown in Attachment1 I have the power spectrum of the 32-bit output and the 64-bit output as well as the power spectrum of the two subtracted time series as well as the subtracted power spectra of both. unfortunately, all of them follow the same general shape of the raw output of the filter. 

I would not expect quantization noise to follow the shape of the filter. I would instead expect it to be more uniform. If anything I would expect the quantization noise to increase with frequency. If a high-frequency signal is run through a filter that has high quantization noise then it will severely degrade: i.e. higher quantization noise. 

This is one reason why I am confused by what I am seeing here. In all cases including feeding the same and different white noise into both filters, I have found that the calculated quantization noise is proportional to the response of the filter. this seems wrong to me so I will continue to play around with it to see if I can gain any intuition about what might be happening.

Attachment 1: DeltaNoiseSpectrum.pdf
DeltaNoiseSpectrum.pdf
  16362   Mon Sep 27 17:04:43 2021 ranaSummaryComputersQuantization Noise Calculation Summary

I suggest that you

  1. change the corner frequency to 10 Hz as I suggested last week. This filter, as it is, is going to give you trouble.
  2. Put in a sine wave at 3.4283 Hz with an amplitude of 1, rather than white noise. In this way, its not necessary to do any subtraction. Just make PSD of the output of each filter.
  3. Be careful about window length and window function. If you don't do this carefully, your PSD will be polluted by window bleeding.

 

  16363   Tue Sep 28 16:31:52 2021 PacoSummaryCalibrationXARM OLTF (calibration) at 55.511 Hz

[anchal, paco]

Here is a demonstration of the methods leading to the single (X)arm calibration with its budget uncertainty. The steps towards this measurement are the following:

  1. We put a single line excitation through the C1:SUS-ETMX_LSC_EXC at 55.511 Hz, amp = 1 counts, gain = 300 (ramptime=10 s).
  2. With the arm locked, we grab a long timeseries of the C1:LSC-XARM_IN1_DQ (error point) and C1:SUS-ETMX_LSC_OUT_DQ (control point) channels.
  3. We assume the single arm loop to have the four blocks shown in Attachment #1, A (actuator + sus), plant (mainly the cavity pole), D (detection + electronics), and K (digital control).
    1. At this point, Anchal made a model of the single arm loop including the appropriate filter coefficients and other parameters. See Attachments #2-3 for the split and total model TFs.
    2. Our line would actually probe a TF from point b (error point) to point d (control point). We multiplied our measurement with open loop TF from b to d from model to get complete OLTF.
    3. Our initial estimate from documents and elog made overall loop shape correct but it was off by an overall gain factor. This could be due to wrong assumption on RFPD transimpedance or analog gains of AA or whitening filters. We have corrected for this factor in the RFPD transimpedance, but this needs to be checked (if we really care).
  4. We demodulate decimated timeseries (final sampling rate ~ 2.048 kHz) and I & Q for both the b and d signals. From this and our model for K, we estimate the OLTF. Attachment #4 shows timeseries for magnitude and phase.
  5. Finally, we compute the ASD for the OLTF magnitude. We plot it in Attachment #5 together with the ASD of the XARM transmission (C1:LSC-TRX_OUT_DQ) times the OLTF to estimate the optical gain noise ASD (this last step was a quick attempt at budgeting the calibration noise).
    1. For each ASD we used N = 24 averages, from which we estimate rms (statistical) uncertainties which are depicted by error bands (\pm \sigma) around the lines.

** Note: We ran the same procedure using dtt (diaggui) to validate our estimates at every point, as well as check our SNR in b and d before taking the ~3.5 hours of data.

Attachment 1: OLTF_Calibration_Scheme.jpg
OLTF_Calibration_Scheme.jpg
Attachment 2: XARM_POX_Lock_Model_TF.pdf
XARM_POX_Lock_Model_TF.pdf
Attachment 3: XARM_OLTF_Total_Model.pdf
XARM_OLTF_Total_Model.pdf
Attachment 4: XARM_OLTF_55p511_Hz_timeseries.pdf
XARM_OLTF_55p511_Hz_timeseries.pdf
Attachment 5: Gmag_55p511_Hz_ASD.pdf
Gmag_55p511_Hz_ASD.pdf
  16365   Wed Sep 29 17:10:09 2021 AnchalSummaryCDSc1teststand problems summary

[anchal, ian]

We went and collected some information for the overlords to fix the c1teststand DAQ network issue.


  • from c1teststand, c1bhd and c1sus2 computers were not accessible through ssh. (No route to host). So we restarted both the computers (the I/O chassis were ON).
  • After the computers restarted, we were able to ssh into c1bhd and c1sus, ad we ran rtcds start c1x06 and rtcds start c1x07.
  • The first page in attachment shows the screenshot of GDS_TP screens of the IOP models after this step.
  • Then we started teh user models by running rtcds start c1bhd and rtcds start c1su2.
  • The second page shows the screenshot of GDS_TP screens. You can notice that DAQ status is red in all the screens and the DC statuses are blank.
  • So we checked if daqd_ services are running in the fb computer. They were not. So we started them all by sudo systemctl start daqd_*.
  • Third page shows the status of all services after this step. the daqd_dc.service remained at failed state.
  • open-mx_stream.service was not even loaded in fb. We started it by running sudo systemctl start open-mx_stream.service.
  • The fourth page shows the status of this service. It started without any errors.
  • However, when we went to check the status of mx_stream.service in c1bhd and c1sus2, they were not loaded and we we tried to start them, they showed failed state and kept trying to start every 3 seconds without success. (See page 5 and 6).
  • Finally, we also took a screenshot of timedatectl command output on the three computers fb, c1bhd, and c1sus2 to show that their times were not synced at all.
  • The ntp service is running on fb but it probably does not have access to any of the servers it is following.
  • The timesyncd on c1bhd and c1sus2 (FE machines) is also running but showing status 'Idle' which suggested they are unable to find the ntp signal from fb.
  • I believe this issue is similar to what jamie ficed in the fb1 on martian network in 40m/16302. Since the fb on c1teststand network was cloned before this fix, it might have this dysfunctional ntp as well.

We would try to get internet access to c1teststand soon. Meanwhile, someone with more experience and knowledge should look into this situation and try to fix it. We need to test the c1teststand within few weeks now.

Attachment 1: c1teststand_issues_summary.pdf
c1teststand_issues_summary.pdf c1teststand_issues_summary.pdf c1teststand_issues_summary.pdf c1teststand_issues_summary.pdf c1teststand_issues_summary.pdf c1teststand_issues_summary.pdf c1teststand_issues_summary.pdf
  16366   Thu Sep 30 11:46:33 2021 Ian MacMillanSummaryComputersQuantization Noise Calculation Summary

First and foremost I have the updated bode plot with the mode moved to 10 Hz. See Attachment 1. Note that the comparison measurement is a % difference whereas in the previous bode plot it was just the difference. I also wrapped the phase so that jumps from -180 to 180 are moved down. This eliminates massive jumps in the % difference. 

Next, I have two comparison plots: 32 bit and 64bit. As mentioned above I moved the mode to 10 Hz and just excited both systems at 3.4283Hz with an amplitude of 1. As we can see on the plot the two models are practically the same when using 64bits. With the 32bit system, we can see that the noise in the IIR filter is much greater than in the State-space model at frequencies greater than our mode.

Note about windowing and averaging: I used a Hanning window with averaging over 4 neighbor points. I came to this number after looking at the results with less averaging and more averaging. In the code, this can be seen as nperseg=num_samples/4 which is then fed into signal.welch

Attachment 1: SS-IIR-Bode.pdf
SS-IIR-Bode.pdf
Attachment 2: PSD_32bit.pdf
PSD_32bit.pdf
Attachment 3: PSD_64bit.pdf
PSD_64bit.pdf
  16367   Thu Sep 30 14:09:37 2021 AnchalSummaryCDSNew way to ssh into c1teststand

Late elog, original time Wed Sep 29 14:09:59 2021

We opened a new port (22220) in the router to the martian subnetwork which is forwarded to port 22 on c1teststand (192.168.113.245) allowing direct ssh access to c1teststand computer from the outside world using:

                                                                       

                                                                                    
 

Checkout this wiki page for unredadcted info.

  16369   Thu Sep 30 18:04:31 2021 PacoSummaryCalibrationXARM OLTF (calibration) with three lines

[anchal, paco]

We repeated the same procedure as before, but with 3 different lines at 55.511, 154.11, and 1071.11 Hz. We overlay the OLTF magnitudes and phases with our latest model (which we have updated with Koji's help) and include the rms uncertainties as errorbars in Attachment #1.

We also plot the noise ASDs of calibrated OLTF magnitudes at the line frequencies in Attachment #2. These curves are created by calculating power spectral density of timeseries of OLTF values at the line frequencies generated by demodulated XARM_IN and ETMX_LSC_OUT signals. We have overlayed the TRX noise spectrum here as an attempt to see if we can budget the noise measured in values of G to the fluctuation in optical gain due to changing power in the arms. We multiplied the the transmission ASD with the value of OLTF at those frequencies as the transfger function from normalized optical gain to the total transfer function value.

It is weird that the fluctuations in transmission power at 1 mHz always crosses the total noise in the OLTF value in all calibration lines. This could be an artificat of our data analysis though.

Even if the contribution of the fluctuating power is correct, there is remaining excess noise in the OLTF to be budgeted.

Attachment 1: XARM_OLTF_Model_and_Meas.pdf
XARM_OLTF_Model_and_Meas.pdf XARM_OLTF_Model_and_Meas.pdf
Attachment 2: Gmag_ASD_nb_withTRX.pdf
Gmag_ASD_nb_withTRX.pdf Gmag_ASD_nb_withTRX.pdf Gmag_ASD_nb_withTRX.pdf
  16371   Fri Oct 1 14:25:27 2021 yehonathanSummarySUSPRM and BS Angular Actuation transfer function magnitude measurements

{Paco, Yehonathan, Hang}

We measured the sensing PRMI sensing matrix. Attachment 1 shows the results, the magnitude of the response is not calibrated. The orthogonality between PRCL and MICH is still bad (see previous measurement for reference).

Hang suggested that since MICH actuation with BS and PRM is not trivial (0.5*BS - 0.34*PRM) and since PRCL is so sensitive to PRM movement there might be a leakage to PRCL when we are actuating on MICH. So there may be a room to tune the PRM coefficient in the MICH output matrix.

Attachment 2 shows the sensing matrix after we changed the MICH->PRM coefficient in the OSC output matrix to -0.1.

It seems like it made things a little bit better but not much and also there is a huge uncertainty in the MICH sensing.

Attachment 1: MICH_PRM_-0.34.png
MICH_PRM_-0.34.png
Attachment 2: MICH_PRM_-0.1.png
MICH_PRM_-0.1.png
  16372   Mon Oct 4 11:05:44 2021 AnchalSummaryCDSc1teststand problems summary

[Anchal, Paco]

We tried to fix the ntp synchronization in c1teststand today by repeating the steps listed in 40m/16302. Even though teh cloned fb1 now has the exact same package version, conf & service files, and status, the FE machines (c1bhd and c1sus2) fail to sync to the time. the timedatectl shows the same stauts 'Idle'. We also, dug bit deeper into the error messages of daq_dc on cloned fb1 and mx_stream on FE machines and have some error messages to report here.


Attempt on fixing the ntp

  • We copied the ntp package version 1:4.2.6 deb file from /var/cache/apt/archives/ntp_1%3a4.2.6.p5+dfsg-7+deb8u3_amd64.deb on the martian fb1 to the cloned fb1 and ran.
    controls@fb1:~ 0$ sudo dbpg -i ntp_1%3a4.2.6.p5+dfsg-7+deb8u3_amd64.deb
  • We got error messages about missing dependencies of libopts25 and libssl1.1. We downloaded oldoldstable jessie versions of these packages from here and here. We ensured that these versions are higher than the required versions for ntp. We installed them with:
    controls@fb1:~ 0$ sudo dbpg -i libopts25_5.18.12-3_amd64.deb 
    controls@fb1:~ 0$ sudo dbpg -i libssl1.1_1.1.0l-1~deb9u4_amd64.deb
  • Then we installed the ntp package as described above. It asked us if we want to keep the configuration file, we pressed Y.
  • However, we decided to make the configuration and service files exactly same as martian fb1 to make it same in cloned fb1. We copied /etc/ntp.conf and /etc/systemd/system/ntp.service files from martian fb1 to cloned fb1 in the same positions. Then we enabled ntp, reloaded the daemon, and restarted ntp service:
    controls@fb1:~ 0$ sudo systemctl enable ntp
    controls@fb1:~ 0$ sudo systemctl daemon-reload
    controls@fb1:~ 0$ sudo systemctl restart ntp
  • But ofcourse, since fb1 doesn't have internet access, we got some errors in status of the ntp.service:
    controls@fb1:~ 0$ sudo systemctl status ntp
    ● ntp.service - NTP daemon (custom service)
       Loaded: loaded (/etc/systemd/system/ntp.service; enabled)
       Active: active (running) since Mon 2021-10-04 17:12:58 UTC; 1h 15min ago
     Main PID: 26807 (code=exited, status=0/SUCCESS)
       CGroup: /system.slice/ntp.service
               ├─30408 /usr/sbin/ntpd -p /var/run/ntpd.pid -g -u 105:107
               └─30525 /usr/sbin/ntpd -p /var/run/ntpd.pid -g -u 105:107
    
    Oct 04 17:48:42 fb1 ntpd_intres[30525]: host name not found: 2.debian.pool.ntp.org
    Oct 04 17:48:52 fb1 ntpd_intres[30525]: host name not found: 3.debian.pool.ntp.org
    Oct 04 18:05:05 fb1 ntpd_intres[30525]: host name not found: 0.debian.pool.ntp.org
    Oct 04 18:05:15 fb1 ntpd_intres[30525]: host name not found: 1.debian.pool.ntp.org
    Oct 04 18:05:25 fb1 ntpd_intres[30525]: host name not found: 2.debian.pool.ntp.org
    Oct 04 18:05:35 fb1 ntpd_intres[30525]: host name not found: 3.debian.pool.ntp.org
    Oct 04 18:21:48 fb1 ntpd_intres[30525]: host name not found: 0.debian.pool.ntp.org
    Oct 04 18:21:58 fb1 ntpd_intres[30525]: host name not found: 1.debian.pool.ntp.org
    Oct 04 18:22:08 fb1 ntpd_intres[30525]: host name not found: 2.debian.pool.ntp.org
    Oct 04 18:22:18 fb1 ntpd_intres[30525]: host name not found: 3.debian.pool.ntp.org
  • But the ntpq command is giving the saem output as given by ntpq comman in martian fb1 (except for the source servers), that the broadcasting is happening in the same manner:
    controls@fb1:~ 0$ ntpq -p
         remote           refid      st t when poll reach   delay   offset  jitter
    ==============================================================================
     192.168.123.255 .BCST.          16 u    -   64    0    0.000    0.000   0.000
    
  • On the FE machines side though, the systemd-timesyncd are still unable to read the time signal from fb1 and show the status as idle:
    controls@c1bhd:~ 3$ timedatectl
          Local time: Mon 2021-10-04 18:34:38 UTC
      Universal time: Mon 2021-10-04 18:34:38 UTC
            RTC time: Mon 2021-10-04 18:34:38
           Time zone: Etc/UTC (UTC, +0000)
         NTP enabled: yes
    NTP synchronized: no
     RTC in local TZ: no
          DST active: n/a
    controls@c1bhd:~ 0$ systemctl status systemd-timesyncd -l
    ● systemd-timesyncd.service - Network Time Synchronization
       Loaded: loaded (/lib/systemd/system/systemd-timesyncd.service; enabled)
       Active: active (running) since Mon 2021-10-04 17:21:29 UTC; 1h 13min ago
         Docs: man:systemd-timesyncd.service(8)
     Main PID: 244 (systemd-timesyn)
       Status: "Idle."
       CGroup: /system.slice/systemd-timesyncd.service
               └─244 /lib/systemd/systemd-timesyncd
  • So the time synchronization is still not working. We expected the FE machined to just synchronize to fb1 even though it doesn't have any upstream ntp server to synchronize to. But that didn't happen.
  • I'm (Anchal) working on getting internet access to c1teststand computers.

Digging into mx_stream/daqd_dc errors:

  • We went and changed the Restart fileld in /etc/systemd/system/daqd_dc.service on cloned fb1 to 2. This allows the service to fail and stop restarting after two attempts. This allows us to see the real error message instead of the systemd error message that the service is restarting too often. We got following:
    controls@fb1:~ 3$ sudo systemctl status daqd_dc -l
    ● daqd_dc.service - Advanced LIGO RTS daqd data concentrator
       Loaded: loaded (/etc/systemd/system/daqd_dc.service; enabled)
       Active: failed (Result: exit-code) since Mon 2021-10-04 17:50:25 UTC; 22s ago
      Process: 715 ExecStart=/usr/bin/daqd_dc_mx -c /opt/rtcds/caltech/c1/target/daqd/daqdrc.dc (code=exited, status=1/FAILURE)
     Main PID: 715 (code=exited, status=1/FAILURE)
    
    Oct 04 17:50:24 fb1 systemd[1]: Started Advanced LIGO RTS daqd data concentrator.
    Oct 04 17:50:25 fb1 daqd_dc_mx[715]: [Mon Oct  4 17:50:25 2021] Unable to set to nice = -20 -error Unknown error -1
    Oct 04 17:50:25 fb1 daqd_dc_mx[715]: Failed to do mx_get_info: MX not initialized.
    Oct 04 17:50:25 fb1 daqd_dc_mx[715]: 263596
    Oct 04 17:50:25 fb1 systemd[1]: daqd_dc.service: main process exited, code=exited, status=1/FAILURE
    Oct 04 17:50:25 fb1 systemd[1]: Unit daqd_dc.service entered failed state.
    
  • It seemed like the only thing daqd_dc process doesn't like is that mx_stream services are in failed state in teh FE computers. So we did the same process on FE machines to get the real error messages:
    controls@fb1:~ 0$ sudo chroot /diskless/root
    fb1:/ 0#
    fb1:/ 0# sudo nano /etc/systemd/system/mx_stream.service
    fb1:/ 0#
    fb1:/ 0# exit
  • Then I ssh'ed into c1bhd to see the error message on mx_stream service properly.
    controls@c1bhd:~ 0$ sudo systemctl daemon-reload
    controls@c1bhd:~ 0$ sudo systemctl restart mx_stream
    controls@c1bhd:~ 0$ sudo systemctl status mx_stream -l
    ● mx_stream.service - Advanced LIGO RTS front end mx stream
       Loaded: loaded (/etc/systemd/system/mx_stream.service; enabled)
       Active: failed (Result: exit-code) since Mon 2021-10-04 17:57:20 UTC; 24s ago
      Process: 11832 ExecStart=/etc/mx_stream_exec (code=exited, status=1/FAILURE)
     Main PID: 11832 (code=exited, status=1/FAILURE)
    
    Oct 04 17:57:20 c1bhd systemd[1]: Starting Advanced LIGO RTS front end mx stream...
    Oct 04 17:57:20 c1bhd systemd[1]: Started Advanced LIGO RTS front end mx stream.
    Oct 04 17:57:20 c1bhd mx_stream_exec[11832]: send len = 263596
    Oct 04 17:57:20 c1bhd mx_stream_exec[11832]: OMX: Failed to find peer index of board 00:00:00:00:00:00 (Peer Not Found in the Table)
    Oct 04 17:57:20 c1bhd mx_stream_exec[11832]: mx_connect failed Nic ID not Found in Peer Table
    Oct 04 17:57:20 c1bhd mx_stream_exec[11832]: c1x06_daq mmapped address is 0x7f516a97a000
    Oct 04 17:57:20 c1bhd mx_stream_exec[11832]: c1bhd_daq mmapped address is 0x7f516697a000
    Oct 04 17:57:20 c1bhd systemd[1]: mx_stream.service: main process exited, code=exited, status=1/FAILURE
    Oct 04 17:57:20 c1bhd systemd[1]: Unit mx_stream.service entered failed state.
    
  • c1sus2 shows the same error. I'm not sure I understand these errors at all. But they seem to have nothing to do with timing issuessurprise!

As usual, some help would be helpful

  16374   Mon Oct 4 16:00:57 2021 YehonathanSummarySUSPRM and BS Angular Actuation transfer function magnitude measurements

{Yehonathan, Anchel}

In an attempt to fix the actuation of the PRMI DOFs we set to modify the output matrix of the BS and PRM such that the response of the coils will be similar to each other as much as possible.

To do so, we used the responses at a single frequency from the previous measurement to infer the output matrix coefficients that will equilize the OpLev responses (arbitrarily making the LL coil as a reference). This corrected the imbalance in BS almost completely while it didn't really work for PRM (see attachment 1).

The new output matrices are shown in attachment 2-3.

Attachment 1: BS_PRM_ANG_ACT_TF_20211004.pdf
BS_PRM_ANG_ACT_TF_20211004.pdf BS_PRM_ANG_ACT_TF_20211004.pdf BS_PRM_ANG_ACT_TF_20211004.pdf BS_PRM_ANG_ACT_TF_20211004.pdf
Attachment 2: BS_out_mat_20211004.txt
9.839999999999999858e-01 8.965770586285104482e-01 9.486710352885977526e-01 3.099999999999999978e-01
1.016000000000000014e+00 9.750242104232501594e-01 -9.291967546765563801e-01 3.099999999999999978e-01
9.839999999999999858e-01 -1.086765190351774768e+00 1.009798093279114628e+00 3.099999999999999978e-01
1.016000000000000014e+00 -1.031706735496689786e+00 -1.103142995587099939e+00 3.099999999999999978e-01
0.000000000000000000e+00 0.000000000000000000e+00 0.000000000000000000e+00 1.000000000000000000e+00
Attachment 3: PRM_out_mat_20211004.txt
1.000000000000000000e+00 1.033455230230304611e+00 9.844796282226820905e-01 0.000000000000000000e+00
1.000000000000000000e+00 9.342329554807877745e-01 -1.021296201828568506e+00 0.000000000000000000e+00
1.000000000000000000e+00 -1.009214777246558503e+00 9.965113815550634691e-01 0.000000000000000000e+00
1.000000000000000000e+00 -1.020129700278567197e+00 -9.973560027273553619e-01 0.000000000000000000e+00
0.000000000000000000e+00 0.000000000000000000e+00 0.000000000000000000e+00 1.000000000000000000e+00
  16375   Mon Oct 4 16:10:09 2021 ranaSummarySUSPRM and BS Angular Actuation transfer function magnitude measurements

not sure that this is necessary. If you look at teh previous entries Gautam made on this topic, it is clear that the BS/PRM PRMI matrix is snafu, whereas the ITM PRMI matrix is not.

Is it possible that the ~5% coil imbalance of the BS/PRM can explain the observed sensing matrix? If not, then there is no need to balance these coils.

  16376   Mon Oct 4 18:00:16 2021 KojiSummaryCDSc1teststand problems summary

I don't know anything about mx/open-mx, but you also need open-mx,don't you?


controls@c1ioo:~ 0$ systemctl status *mx*
● open-mx.service - LSB: starts Open-MX driver
   Loaded: loaded (/etc/init.d/open-mx)
   Active: active (running) since Wed 2021-09-22 11:54:39 PDT; 1 weeks 5 days ago
  Process: 470 ExecStart=/etc/init.d/open-mx start (code=exited, status=0/SUCCESS)
   CGroup: /system.slice/open-mx.service
           └─620 /opt/3.2.88-csp/open-mx-1.5.4/bin/fma -d

● mx_stream.service - Advanced LIGO RTS front end mx stream
   Loaded: loaded (/etc/systemd/system/mx_stream.service; enabled)
   Active: active (running) since Wed 2021-09-22 12:08:00 PDT; 1 weeks 5 days ago
 Main PID: 5785 (mx_stream)
   CGroup: /system.slice/mx_stream.service
           └─5785 /usr/bin/mx_stream -e 0 -r 0 -w 0 -W 0 -s c1x03 c1ioo c1als c1omc -d fb1:0

 

  16381   Tue Oct 5 17:58:52 2021 AnchalSummaryCDSc1teststand problems summary

open-mx service is running successfully on the fb1(clone), c1bhd and c1sus.

Quote:

I don't know anything about mx/open-mx, but you also need open-mx,don't you?


  16382   Tue Oct 5 18:00:53 2021 AnchalSummaryCDSc1teststand time synchronization working now

Today I got a new router that I used to connect the c1teststand, fb1 and chiara. I was able to see internet access in c1teststand and fb1, but not in chiara. I'm not sure why that is the case.

The good news is that the ntp server on fb1(clone) is working fine now and both FE computers, c1bhd and c1sus2 are succesfully synchronized to the fb1(clone) ntpserver. This resolves any possible timing issues in this DAQ network.

On running the IOP and user models however, I see the same errors are mentioned in 40m/16372. Something to do with:

Oct 06 00:47:56 c1sus2 mx_stream_exec[21796]: OMX: Failed to find peer index of board 00:00:00:00:00:00 (Peer Not Found in the Table)
Oct 06 00:47:56 c1sus2 mx_stream_exec[21796]: mx_connect failed Nic ID not Found in Peer Table
Oct 06 00:47:56 c1sus2 mx_stream_exec[21796]: c1x07_daq mmapped address is 0x7fa4819cc000
Oct 06 00:47:56 c1sus2 mx_stream_exec[21796]: c1su2_daq mmapped address is 0x7fa47d9cc000


Thu Oct 7 17:04:31 2021

I fixed the issue of chiara not getting internet. Now c1teststand, fb1 and chiara, all have internet connections. It was the issue of default gateway and interface and findiing the DNS. I have found the correct settings now.

  16383   Tue Oct 5 20:04:22 2021 PacoSummarySUSPRM and BS Angular Actuation transfer function magnitude measurements

[Paco, Rana]

We had a look at the BS actuation. Along the way we created a couple of issues that we fixed. A summary is below.

  1. First, we locked MICH. While doing this, we used the /users/Templates/ndscope/LSC/MICH.yml ndscope template to monitor some channels. I edited the yaml file to look at C1:LSC-ASDC_OUT_DQ instead of the REFL_DC. Rana pointed out that the C1:LSC-MICH_OUT_DQ (MICH control point) had a big range (~ 5000 counts rms) and this should not be like that.
  2. We tried to investigate the aforementioned thing by looking at the whitening / uwhitening filters but all the slow epics channels where "white" on the medm screen. Looking under CDS/slow channel monitors, we realized that both c1iscaux and c1auxey were weird, so we tried telnet to c1iscaux without success. Therefore, we followed the recommended wiki procedure of hard rebooting this machine. While inside the lab and looking for this machine, we touched things around the 'rfpd' rack and once we were back in the control room, we couldn't see any light on the AS port camera. But the whitening filter medm screens were back up.
  3. While rana ssh'd into c1auxey to investigate about its status, and burtrestored the c1iscaux channels, we looked at trends to figure out if anything had changed (for example TT1 or TT2) but this wasn't the case. We decided to go back inside to check the actual REFL beams and noticed it was grossly misaligned (clipping)... so we blamed it on the TTs and again, went around and moved some stuff around the 'rfpd' rack. We didn't really connect or disconnect anything, but once we were back in the control room, light was coming from the AS port again. This is a weird mystery and we should systematically try to repeat this and fix the actual issue.
  4. We restored the MICH, and returned to BS actuation problems. Here, we essentially devised a scheme to inject noise at 310.97 Hz and 313.74. The choice is twofold, first it lies outside the MICH loop UGF (~150 Hz), and second, it matches the sensing matrix OSC frequencies, so it's more appropriate for a comparison.
  5. We injected two lines using the BS SUS LOCKIN1 and LOCKIN2 oscilators so we can probe two coils at once, with the LSC loop closed, and read back using the C1:LSC-MICH_IN1_DQ channel. We excited with an amplitude of 1234.0 counts and 1254 counts respectively (to match the ~ 2 % difference in frequency) and noted that the magnitude response in UR was 10% larger than UL, LL, and LR which were close to each other at the 2% level.

[Paco]

After rana left, I did a second pass at the BS actuation. I took TF measurements at the oscilator frequencies noted above using diaggui, and summarize the results below:

TF UL (310.97 Hz) UR (313.74 Hz) LL (310.97 Hz) LR (313.74 Hz)
Magnitude (dB) 93.20 92.20 94.27 93.85
Phase (deg) -128.3 -127.9 -128.4 -127.5

This procedure should be done with PRM as well and using the PRCL instead of MICH.

  16385   Wed Oct 6 15:39:29 2021 AnchalSummarySUSPRM and BS Angular Actuation transfer function magnitude measurements

Note that your tests were done with the output matrix for BS and PRM in the compensated state as done in 40m/16374. The changes made there were supposed to clear out any coil actuation imbalance in the angular degrees of freedom.

  16391   Mon Oct 11 17:31:25 2021 AnchalSummaryCDSFixed mounting of mx devices in fb. daqd_dc is running now.
 
 

However, lspci | grep 'Myri' shows following output on both computers:

controls@fb1:/dev 0$ lspci | grep 'Myri'
02:00.0 Ethernet controller: MYRICOM Inc. Myri-10G Dual-Protocol NIC (rev 01)

Which means that the computer detects the card on PCie slot.

 

I tried to add this to /etc/rc.local to run this script at every boot, but it did not work. So for now, I'll just manually do this step everytime. Once the devices are loaded, we get:

controls@fb1:/etc 0$ ls /dev/*mx*
/dev/mx0  /dev/mx4  /dev/mxctl   /dev/mxp2  /dev/mxp6         /dev/ptmx
/dev/mx1  /dev/mx5  /dev/mxctlp  /dev/mxp3  /dev/mxp7
/dev/mx2  /dev/mx6  /dev/mxp0    /dev/mxp4  /dev/open-mx
/dev/mx3  /dev/mx7  /dev/mxp1    /dev/mxp5  /dev/open-mx-raw

The, restarting all daqd_ processes, I found that daqd_dc was running succesfully now. Here is the status:

controls@fb1:/etc 0$ sudo systemctl status daqd_* -l
● daqd_dc.service - Advanced LIGO RTS daqd data concentrator
   Loaded: loaded (/etc/systemd/system/daqd_dc.service; enabled)
   Active: active (running) since Mon 2021-10-11 17:48:00 PDT; 23min ago
 Main PID: 2308 (daqd_dc_mx)
   CGroup: /daqd.slice/daqd_dc.service
           ├─2308 /usr/bin/daqd_dc_mx -c /opt/rtcds/caltech/c1/target/daqd/daqdrc.dc
           └─2370 caRepeater

Oct 11 17:48:07 fb1 daqd_dc_mx[2308]: mx receiver 006 thread priority error Operation not permitted[Mon Oct 11 17:48:06 2021]
Oct 11 17:48:07 fb1 daqd_dc_mx[2308]: mx receiver 005 thread put on CPU 0
Oct 11 17:48:07 fb1 daqd_dc_mx[2308]: [Mon Oct 11 17:48:06 2021] [Mon Oct 11 17:48:06 2021] mx receiver 006 thread put on CPU 0
Oct 11 17:48:07 fb1 daqd_dc_mx[2308]: mx receiver 007 thread put on CPU 0
Oct 11 17:48:07 fb1 daqd_dc_mx[2308]: [Mon Oct 11 17:48:06 2021] mx receiver 003 thread - label dqmx003 pid=2362
Oct 11 17:48:07 fb1 daqd_dc_mx[2308]: [Mon Oct 11 17:48:06 2021] mx receiver 003 thread priority error Operation not permitted
Oct 11 17:48:07 fb1 daqd_dc_mx[2308]: [Mon Oct 11 17:48:06 2021] mx receiver 003 thread put on CPU 0
Oct 11 17:48:07 fb1 daqd_dc_mx[2308]: warning:regcache incompatible with malloc
Oct 11 17:48:07 fb1 daqd_dc_mx[2308]: [Mon Oct 11 17:48:06 2021] EDCU has 410 channels configured; first=0
Oct 11 17:49:06 fb1 daqd_dc_mx[2308]: [Mon Oct 11 17:49:06 2021] ->4: clear crc

● daqd_fw.service - Advanced LIGO RTS daqd frame writer
   Loaded: loaded (/etc/systemd/system/daqd_fw.service; enabled)
   Active: active (running) since Mon 2021-10-11 17:48:01 PDT; 23min ago
 Main PID: 2318 (daqd_fw)
   CGroup: /daqd.slice/daqd_fw.service
           └─2318 /usr/bin/daqd_fw -c /opt/rtcds/caltech/c1/target/daqd/daqdrc.fw

Oct 11 17:48:09 fb1 daqd_fw[2318]: [Mon Oct 11 17:48:09 2021] [Mon Oct 11 17:48:09 2021] Producer thread - label dqproddbg pid=2440
Oct 11 17:48:09 fb1 daqd_fw[2318]: Producer crc thread priority error Operation not permitted
Oct 11 17:48:09 fb1 daqd_fw[2318]: [Mon Oct 11 17:48:09 2021] [Mon Oct 11 17:48:09 2021] Producer crc thread put on CPU 0
Oct 11 17:48:09 fb1 daqd_fw[2318]: Producer thread priority error Operation not permitted
Oct 11 17:48:09 fb1 daqd_fw[2318]: [Mon Oct 11 17:48:09 2021] Producer thread put on CPU 0
Oct 11 17:48:09 fb1 daqd_fw[2318]: [Mon Oct 11 17:48:09 2021] Producer thread - label dqprod pid=2434
Oct 11 17:48:09 fb1 daqd_fw[2318]: [Mon Oct 11 17:48:09 2021] Producer thread priority error Operation not permitted
Oct 11 17:48:09 fb1 daqd_fw[2318]: [Mon Oct 11 17:48:09 2021] Producer thread put on CPU 0
Oct 11 17:48:10 fb1 daqd_fw[2318]: [Mon Oct 11 17:48:10 2021] Minute trender made GPS time correction; gps=1318034906; gps%60=26
Oct 11 17:49:09 fb1 daqd_fw[2318]: [Mon Oct 11 17:49:09 2021] ->3: clear crc

● daqd_rcv.service - Advanced LIGO RTS daqd testpoint receiver
   Loaded: loaded (/etc/systemd/system/daqd_rcv.service; enabled)
   Active: active (running) since Mon 2021-10-11 17:48:00 PDT; 23min ago
 Main PID: 2311 (daqd_rcv)
   CGroup: /daqd.slice/daqd_rcv.service
           └─2311 /usr/bin/daqd_rcv -c /opt/rtcds/caltech/c1/target/daqd/daqdrc.rcv

Oct 11 17:50:21 fb1 daqd_rcv[2311]: Creating C1:DAQ-NDS0_C1X07_CRC_SUM
Oct 11 17:50:21 fb1 daqd_rcv[2311]: Creating C1:DAQ-NDS0_C1BHD_STATUS
Oct 11 17:50:21 fb1 daqd_rcv[2311]: Creating C1:DAQ-NDS0_C1BHD_CRC_CPS
Oct 11 17:50:21 fb1 daqd_rcv[2311]: Creating C1:DAQ-NDS0_C1BHD_CRC_SUM
Oct 11 17:50:21 fb1 daqd_rcv[2311]: Creating C1:DAQ-NDS0_C1SU2_STATUS
Oct 11 17:50:21 fb1 daqd_rcv[2311]: Creating C1:DAQ-NDS0_C1SU2_CRC_CPS
Oct 11 17:50:21 fb1 daqd_rcv[2311]: Creating C1:DAQ-NDS0_C1SU2_CRC_SUM
Oct 11 17:50:21 fb1 daqd_rcv[2311]: Creating C1:DAQ-NDS0_C1OM[Mon Oct 11 17:50:21 2021] Epics server started
Oct 11 17:50:24 fb1 daqd_rcv[2311]: [Mon Oct 11 17:50:24 2021] Minute trender made GPS time correction; gps=1318035040; gps%120=40
Oct 11 17:51:21 fb1 daqd_rcv[2311]: [Mon Oct 11 17:51:21 2021] ->3: clear crc

Now, even before starting teh FE models, I see DC status as ox2bad in the CDS screens of the IOP and user models. The mx_stream service remains in a failed state at teh FE machines and remain the same even after restarting the service.

controls@c1sus2:~ 0$ sudo systemctl status mx_stream -l
● mx_stream.service - Advanced LIGO RTS front end mx stream
   Loaded: loaded (/etc/systemd/system/mx_stream.service; enabled)
   Active: failed (Result: exit-code) since Mon 2021-10-11 17:50:26 PDT; 15min ago
  Process: 382 ExecStart=/etc/mx_stream_exec (code=exited, status=1/FAILURE)
 Main PID: 382 (code=exited, status=1/FAILURE)

Oct 11 17:50:25 c1sus2 systemd[1]: Starting Advanced LIGO RTS front end mx stream...
Oct 11 17:50:25 c1sus2 systemd[1]: Started Advanced LIGO RTS front end mx stream.
Oct 11 17:50:25 c1sus2 mx_stream_exec[382]: Failed to open endpoint Not initialized
Oct 11 17:50:26 c1sus2 systemd[1]: mx_stream.service: main process exited, code=exited, status=1/FAILURE
Oct 11 17:50:26 c1sus2 systemd[1]: Unit mx_stream.service entered failed state.

But  if I restart the mx_stream service before starting the rtcds models, the mx-stream service starts succesfully:

controls@c1sus2:~ 0$ sudo systemctl restart mx_stream
controls@c1sus2:~ 0$ sudo systemctl status mx_stream -l
● mx_stream.service - Advanced LIGO RTS front end mx stream
   Loaded: loaded (/etc/systemd/system/mx_stream.service; enabled)
   Active: active (running) since Mon 2021-10-11 18:14:13 PDT; 25s ago
 Main PID: 1337 (mx_stream)
   CGroup: /system.slice/mx_stream.service
           └─1337 /usr/bin/mx_stream -e 0 -r 0 -w 0 -W 0 -s c1x07 c1su2 -d fb1:0

Oct 11 18:14:13 c1sus2 systemd[1]: Starting Advanced LIGO RTS front end mx stream...
Oct 11 18:14:13 c1sus2 systemd[1]: Started Advanced LIGO RTS front end mx stream.
Oct 11 18:14:13 c1sus2 mx_stream_exec[1337]: send len = 263596
Oct 11 18:14:13 c1sus2 mx_stream_exec[1337]: Connection Made

However, the DC status on CDS screens still show 0x2bad. As soon as I start the rtcds model c1x07 (the IOP model for c1sus2), the mx_stream service fails:

controls@c1sus2:~ 0$ sudo systemctl status mx_stream -l
● mx_stream.service - Advanced LIGO RTS front end mx stream
   Loaded: loaded (/etc/systemd/system/mx_stream.service; enabled)
   Active: failed (Result: exit-code) since Mon 2021-10-11 18:18:03 PDT; 27s ago
  Process: 1337 ExecStart=/etc/mx_stream_exec (code=exited, status=1/FAILURE)
 Main PID: 1337 (code=exited, status=1/FAILURE)

Oct 11 18:14:13 c1sus2 systemd[1]: Starting Advanced LIGO RTS front end mx stream...
Oct 11 18:14:13 c1sus2 systemd[1]: Started Advanced LIGO RTS front end mx stream.
Oct 11 18:14:13 c1sus2 mx_stream_exec[1337]: send len = 263596
Oct 11 18:14:13 c1sus2 mx_stream_exec[1337]: Connection Made
Oct 11 18:18:03 c1sus2 mx_stream_exec[1337]: isendxxx failed with status Remote Endpoint Unreachable
Oct 11 18:18:03 c1sus2 mx_stream_exec[1337]: disconnected from the sender
Oct 11 18:18:03 c1sus2 mx_stream_exec[1337]: c1x07_daq mmapped address is 0x7fe3620c3000
Oct 11 18:18:03 c1sus2 mx_stream_exec[1337]: c1su2_daq mmapped address is 0x7fe35e0c3000
Oct 11 18:18:03 c1sus2 systemd[1]: mx_stream.service: main process exited, code=exited, status=1/FAILURE
Oct 11 18:18:03 c1sus2 systemd[1]: Unit mx_stream.service entered failed state.

This shows that the start of rtcds model, causes the fail in mx_stream, possibly due to inability of finding the endpoint on fb1. I've again reached to the edge of my knowledge here. Maybe the fiber optic connection between fb and the network switch that connects to FE is bad, or the connection between switch and FEs is bad.

But we are just one step away from making this work.

 

 

  16392   Mon Oct 11 18:29:35 2021 AnchalSummaryCDSMoving forward?

The teststand has some non-trivial issue with Myrinet card (either software or hardware) which even teh experts are saying they don't remember how to fix it. CDS with mx was iin use more than a decade ago, so it is hard to find support for issues with it now and will be the same in future. We need to wrap up this test procedure one way or another now, so I have following two options moving forward:


Direct integration with main CDS and testing

  • We can just connect the c1sus2 and c1bhd FE computers to martian network directly.
  • We'll have to connect c1sus2 and c1bhd to the optical fiber subnetwork as well.
  • On booting, they would get booted through the exisitng fb1 boot server which seems to work fine for the other 5 FE machines.
  • We can update teh DHCP in chiara and reload it so that we can ssh into these FEs with host names.
  • Hopefully, presence of these computers won't tank the existing CDS even if they  themselves have any issues, as they have no shared memory with other models.
  • If this works, we can do the loop back testing of I/O chassis using the main DAQ network and move on with our upgrade.
  • If this does not work and causes any harm to exisitng CDS network, we can disconnect these computers and go back to existing CDS. Recently, our confidence on rebooting the CDS has increased with the robust performance as some legacy issues were fixed.
  • We'll however, continue to use a CDS which is no more supported by the current LIGO CDS group.

Testing CDS upgrade on teststand

  • From what I could gather, most of the hardware in I/O chassis that I could find, is still used in CDS of LLO and LHO, with their recent tests and documents using the same cards and PCBs.
  • There might be some difference in the DAQ network setup that I need to confirm.
  • I've summarised the current c1teststand hardware on this wiki page.
  • If the latest CDS is backwards compatible with our hardware, we can test the new CDS in teh c1teststand setup without disrupting our main CDS. We'll have ample help and support for this upgrade from the current LIGO CDS group.
  • We can do the loop back testing of the I/O chassis as well.
  • If the upgrade is succesfull in the teststand without many hardware changes, we can upgrade the main CDS of 40m as well, as it has the same hardware as our teststand.
  • Biggest plus point would be that out CDS will be up-to-date and we will be able to take help from CDS group if any trouble occurs.

So these are the two options we have. We should discuss which one to take in the mattermost chat or in upcoming meeting.

  16393   Tue Oct 12 11:32:54 2021 YehonathanSummarySUSPRM and BS Angular Actuation transfer function magnitude measurements

Late submission (From Thursday 10/07):

I measured the PRMI sensing matrix to see if the BS and PRMI output matrices tweaking had any effect.

While doing so, I noticed I made a mistake in the analysis of the previous sensing matrix measurement. It seems that I have used the radar plot function with radians where degrees should have been used (the reason is that the azimuthal uncertainty looked crazy when I used degrees. I still don't know why this is the case with this measurement).

In any case, attachment 1 and 2 show the PRMI radar plots with the modified output matrices and and in the normal state, respectively.

It seems like the output matrix modification didn't do anything but REFL55 has good orthogonality. Problem gone??

Attachment 1: modified_output_matrices_radar_plots.png
modified_output_matrices_radar_plots.png
Attachment 2: normal_output_matrices_radar_plots.png
normal_output_matrices_radar_plots.png
  16394   Tue Oct 12 16:39:52 2021 ranaSummarySUSPRM and BS Angular Actuation transfer function magnitude measurements

should compare side by side with the ITM PRMI radar plots to see if there is a difference. How do your new plots compare with Gautam's plots of PRMI?

  16395   Tue Oct 12 17:10:56 2021 AnchalSummaryCDSSome more information

Chris pointed out some information displaying scripts, that show if the DAQ network is working or not. I thought it would be nice to log this information here as well.

controls@fb1:/opt/mx/bin 0$ ./mx_info
MX Version: 1.2.16
MX Build: controls@fb1:/opt/src/mx-1.2.16 Mon Aug 14 11:06:09 PDT 2017
1 Myrinet board installed.
The MX driver is configured to support a maximum of:
    8 endpoints per NIC, 1024 NICs on the network, 32 NICs per host
===================================================================
Instance #0:  364.4 MHz LANai, PCI-E x8, 2 MB SRAM, on NUMA node 0
    Status:        Running, P0: Link Up
    Network:    Ethernet 10G

    MAC Address:    00:60:dd:45:37:86
    Product code:    10G-PCIE-8B-S
    Part number:    09-04228
    Serial number:    423340
    Mapper:        00:60:dd:45:37:86, version = 0x00000000, configured
    Mapped hosts:    3

                                                        ROUTE COUNT
INDEX    MAC ADDRESS     HOST NAME                        P0
-----    -----------     ---------                        ---
   0) 00:60:dd:45:37:86 fb1:0                             1,0
   1) 00:25:90:05:ab:47 c1bhd:0                           1,0
   2) 00:25:90:06:69:c3 c1sus2:0                          1,0

 

controls@c1bhd:~ 1$ /opt/open-mx/bin/omx_info
Open-MX version 1.5.4
 build: root@fb1:/opt/src/open-mx-1.5.4 Tue Aug 15 23:48:03 UTC 2017

Found 1 boards (32 max) supporting 32 endpoints each:
 c1bhd:0 (board #0 name eth1 addr 00:25:90:05:ab:47)
   managed by driver 'igb'

Peer table is ready, mapper is 00:60:dd:45:37:86
================================================
  0) 00:25:90:05:ab:47 c1bhd:0
  1) 00:60:dd:45:37:86 fb1:0
  2) 00:25:90:06:69:c3 c1sus2:0

 

controls@c1sus2:~ 0$ /opt/open-mx/bin/omx_info
Open-MX version 1.5.4
 build: root@fb1:/opt/src/open-mx-1.5.4 Tue Aug 15 23:48:03 UTC 2017

Found 1 boards (32 max) supporting 32 endpoints each:
 c1sus2:0 (board #0 name eth1 addr 00:25:90:06:69:c3)
   managed by driver 'igb'

Peer table is ready, mapper is 00:60:dd:45:37:86
================================================
  0) 00:25:90:06:69:c3 c1sus2:0
  1) 00:60:dd:45:37:86 fb1:0
  2) 00:25:90:05:ab:47 c1bhd:0

These outputs prove that the framebuilder and the FEs are able to see each other in teh DAQ network.


Further, the error that we see when IOP model is started which crashes the mx_stream service on the FE machines (see 40m/16391) :

isendxxx failed with status Remote Endpoint Unreachable

This has been seen earlier when Jamie was troubleshooting the current fb1 in martian network in 40m/11655 in Oct, 2015. Unfortunately, I could not find what Jamie did over a year to fix this issue.

  16396   Tue Oct 12 17:20:12 2021 AnchalSummaryCDSConnected c1sus2 to martian network

I connected c1sus2 to the martian network by splitting the c1sim connection with a 5-way switch. I also ran another ethernet cable from the second port of c1sus2 to the DAQ network switch on 1X7.

Then I logged into chiara and added the following in chiara:/etc/dhcp/dhcpd.conf :

host c1sus2 {
  hardware ethernet 00:25:90:06:69:C2;
  fixed-address 192.168.113.92;
}

And following line in chiara:/var/lib/bind/martian.hosts :

c1sus2          A    192.168.113.92

Note that entires c1bhd is already added in these files, probably during some earlier testing by Gautam or Jon. Then I ran following to restart the dhcp server and nameserver:

~> sudo service bind9 reload
[sudo] password for controls:
 * Reloading domain name service... bind9                                                 [ OK ]
~> sudo service isc-dhcp-server restart
isc-dhcp-server stop/waiting
isc-dhcp-server start/running, process 25764

Now, As I switched on c1sus2 from front panel, it booted over network from fb1 like other FE machines and I was able to login to it by first logging to fb1 and then sshing to c1sus2.

Next, I copied the simulink models and the medm screens of c1x06, xc1x07, c1bhd, c1sus2 from the paths mentioned on this wiki page. I also copied the medm screens from chiara(clone):/opt/rtcds/caltech/c1/medm to martian network chiara in the appropriate places. I have placed the file /opt/rtcds/caltech/c1/medm/teststand_sitemap.adl which can be used to open sitemap for c1bhd and c1sus2 IOP and user models.

Then I logged into c1sus2 (via fb1) and did make, install, start procedure:

controls@c1sus2:~ 0$ rtcds make c1x07
buildd: /opt/rtcds/caltech/c1/rtbuild/release
### building c1x07...
Cleaning c1x07...
Done
Parsing the model c1x07...
Done
Building EPICS sequencers...
Done
Building front-end Linux kernel module c1x07...
Done
RCG source code directory:
/opt/rtcds/rtscore/branches/branch-3.4
The following files were used for this build:
/opt/rtcds/userapps/release/cds/c1/models/c1x07.mdl

Successfully compiled c1x07
***********************************************
Compile Warnings, found in c1x07_warnings.log:
***********************************************
***********************************************
controls@c1sus2:~ 0$ rtcds install c1x07
buildd: /opt/rtcds/caltech/c1/rtbuild/release
### installing c1x07...
Installing system=c1x07 site=caltech ifo=C1,c1
Installing /opt/rtcds/caltech/c1/chans/C1X07.txt
Installing /opt/rtcds/caltech/c1/target/c1x07/c1x07epics
Installing /opt/rtcds/caltech/c1/target/c1x07
Installing start and stop scripts
/opt/rtcds/caltech/c1/scripts/killc1x07
/opt/rtcds/caltech/c1/scripts/startc1x07
sudo: unable to resolve host c1sus2
Performing install-daq
Updating testpoint.par config file
/opt/rtcds/caltech/c1/target/gds/param/testpoint.par
/opt/rtcds/rtscore/branches/branch-3.4/src/epics/util/updateTestpointPar.pl -par_file=/opt/rtcds/caltech/c1/target/gds/param/archive/testpoint_211012_174226.par -gds_node=24 -site_letter=C -system=c1x07 -host=c1sus2
Installing GDS node 24 configuration file
/opt/rtcds/caltech/c1/target/gds/param/tpchn_c1x07.par
Installing auto-generated DAQ configuration file
/opt/rtcds/caltech/c1/chans/daq/C1X07.ini
Installing Epics MEDM screens
Running post-build script

safe.snap exists 
controls@c1sus2:~ 0$ rtcds start c1x07
Cannot start/stop model 'c1x07' on host c1sus2.
controls@c1sus2:~ 4$ rtcds list

controls@c1sus2:~ 0$ 

One can see that even after making and installing, the model c1x07 is not listed as available models in rtcds list. Same is the case for c1sus2 as well. So I could not proceed with testing.

Good news is that nothing that I did affect the current CDS functioning. So we can probably do this testing safely from the main CDS setup.

  16397   Tue Oct 12 23:42:56 2021 KojiSummaryCDSConnected c1sus2 to martian network

Don't you need to add the new hosts to /diskless/root/etc/rtsystab at fb1? --> There looks many elogs talking about editing "rtsystab".

controls@fb1:/diskless/root/etc 0$ cat rtsystab
#
# host    list of control systems to run, starting with IOP
#
c1iscex  c1x01  c1scx c1asx
c1sus     c1x02  c1sus c1mcs c1rfm c1pem
c1ioo     c1x03  c1ioo c1als c1omc
c1lsc    c1x04  c1lsc c1ass c1oaf c1cal c1dnn c1daf
c1iscey  c1x05 c1scy c1asy
#c1test   c1x10  c1tst2

 

  16398   Wed Oct 13 11:25:14 2021 AnchalSummaryCDSRan c1sus2 models in martian CDS. All good!

Three extra steps (when adding new models, new FE):

  • Chris pointed out that the sudo command in c1sus2 is giving error
    sudo: unable to resolve host c1sus2
    
    This error comes in when the computer could not figure out it's own hostname. Since FEs are network booted off the fb1, we need to update the /etc/hosts in /diskless/root everytime we add a new FE.
    controls@fb1:~ 0$ sudo chroot /diskless/root
    fb1:/ 0# sudo nano /etc/hosts
    fb1:/ 0# exit
    
    I added the following line in /etc/hosts file above:
    192.168.113.92  c1sus2 c1sus2.martian
    
    This resolved the issue of sudo giving error. Now, the rtcds make and install steps had no errors mentioned in their outputs.
  • Another thing that needs to be done, as Koji pointed out, is to add the host and models in /etc/rtsystab in /diskless/root of fb:
    controls@fb1:~ 0$ sudo chroot /diskless/root
    fb1:/ 0# sudo nano /etc/rtsystab
    fb1:/ 0# exit
    
    I added the following lines in /etc/rtsystab file above:
    c1sus2   c1x07  c1su2
    
    This told rtcds what models would be available on c1sus2. Now rtcds list is displaying the right models:
    controls@c1sus2:~ 0$ rtcds list
    c1x07
    c1su2
  • The above steps are still not sufficient for the daqd_ processes to know about the new models. This part is supossed to happen automatically, but does not happen in our CDS apparently. So everytime there is a new model, we need to edit the file /opt/rtcds/caltech/c1/target/daqd/master and add following lines to it:
    # Fast Data Channel lists
    # c1sus2
    /opt/rtcds/caltech/c1/chans/daq/C1X07.ini
    /opt/rtcds/caltech/c1/chans/daq/C1SU2.ini
    
    # test point lists
    # c1sus2
    /opt/rtcds/caltech/c1/target/gds/param/tpchn_c1x07.par
    /opt/rtcds/caltech/c1/target/gds/param/tpchn_c1su2.par
    
    I needed to restart the daqd_ processes in  fb1 for them to notice these changes:
    controls@fb1:~ 0$ sudo systemctl restart daqd_*
    
    This finally lit up the status channels of DC in C1X07_GDS_TP.adl and C1SU2_GDS_TP.adl . However the channels C1:DAQ-DC0_C1X07_STATUS and C1:DAQ-DC0_C1SU2_STATUS both have values 0x2bad. This persists on restarting the models. I then just simply restarted teh mx_stream on c1sus2 and boom, it worked! (see attached all green screen, never seen before!)

So now Ian can work on testing the I/O chassis and we would be good to move c1sus2 FE and I/O chassis to 1Y3 after that. I've also done following extra changes:

  • Updated CDS_FE_STATUS medm screen to show the new c1sus2 host.
  • Updated global diag rest script to act on c1xo7 and c1su2 as well.
  • Updated mxstream restart script to act on c1sus2 as well.
Attachment 1: CDS_screens_running.png
CDS_screens_running.png
  16402   Thu Oct 14 13:40:49 2021 YehonathanSummarySUSPRM and BS Angular Actuation transfer function magnitude measurements

Here is a side by side comparison of the PRMI sensing matrix using PRM/BS actuation (attachment 1) and ITMs actuation (attachment 2). The situation looks similar in both cases. That is, good orthogonality on REFL55 and bad seperation in the rest of the RFPDs.

Quote:

should compare side by side with the ITM PRMI radar plots to see if there is a difference. How do your new plots compare with Gautam's plots of PRMI?

 

Attachment 1: BSPRM_Actuation_Radar_plots.png
BSPRM_Actuation_Radar_plots.png
Attachment 2: ITM_Actuation_Radar_plots.png
ITM_Actuation_Radar_plots.png
ELOG V3.1.3-