Fri Sep 21 09:59:09 2018 |
yuki | Configuration | ASC | Y end table upgrade plan |
[Yuki, Gautam]
Attachments #1 is the current setup of AUX Y Green locking and it has to be improved because:
- current efficiency of mode matching is about 50%
- current setup doesn't separate the degrees of freedom of TEM01 with PZT mirrors (the difference of gouy phase between PZT mirrors should be around 90 deg)
- we want to remotely control PZT mirrors for alignment
(Attachments #2 and #3)
About the above two:
One of the example for improvement is just adding a new lens (f=10cm) soon after the doubling crystal. That will make mode matching better (100%) and also make separation better (85 deg) (Attachments #4 and #5). I'm checking whether we have the lens and there is space to set it. And I will measure current power of transmitted main laser in order to confirm the improvement of alignment.
About the last:
I am considering what component is needed.
Reference:
|
14212
|
Sun Sep 23 19:32:23 2018 |
yuki | Configuration | ASC | Y end table upgrade plan |
[ Yuki, Gautam ]
The setup I designed before has abrupt gouy phase shift between two steering mirrors which makes alignment much sensitive. So I designed a new one (Attached #1, #2 and #3). It improves the slope of gouy phase and the difference between steering mirrors is about 100 deg. To install this, we need new lenses: f=100mm, f=200mm, f=-250mm which have 532nm coating. If this setup is OK, I will order them.
There may be a problem: One lens should be put soon after dichroic mirror, but there is little room for fix it. (Attached #4, It will be put where the pedestal is.) Tomorrow we will check this problem again.
And another problem; one steering mirror on the corner of the box is not easy to access. (Attached #5) I have to design a new seup with considering this problem.
Quote: |
One of the example for improvement is just adding a new lens (f=10cm) soon after the doubling crystal. That will make mode matching better (100%) and also make separation better (85 deg) (Attachments #4 and #5). I'm checking whether we have the lens and there is space to set it. And I will measure current power of transmitted main laser in order to confirm the improvement of alignment.
|
|
14214
|
Mon Sep 24 11:09:05 2018 |
yuki | Configuration | ASC | Y end table upgrade plan |
[ Yuki, Steve ]
With Steve's help, we checked a new lens can be set soon after dichroic mirror.
Quote: |
There may be a problem: One lens should be put soon after dichroic mirror, but there is little room for fix it. (Attached #4, It will be put where the pedestal is.) Tomorrow we will check this problem again.
|
|
14216
|
Tue Sep 25 18:08:50 2018 |
yuki | Configuration | ASC | Y end table upgrade plan |
[ Yuki, Gautam ]
We want to remotely control steeing PZT mirrors so its driver is needed. We already have a PZT driver board (D980323-C) and the output voltage is expected to be verified to be in the range 0-100 V DC for input voltages in the range -10 to 10 V DC.
Then I checked to make sure ir perform as we expected. The input signal was supplied using voltage calibrator and the output was monitored using a multimeter.
But it didn't perform well. Some tuning of voltage bias seemed to be needed. I will calculate its transfer function by simulation and check the performance again tommorow. And I found one solder was off so it needs fixing.
Reference:
diagram --> elog 8932
Plan of Action:
- Check PZT driver performs as we expected
- Also check cable, high voltage, PZT mirrors, anti-imaging board
- Obtain calibration factor of PZT mirrors using QPD
- Measure some status value before changing setup (such as tranmitted power of green laser)
- Revise setup after a new lens arrives
- Align the setup and check mode-matching
- Measure status value again and confirm it improves
- (write programming code of making alignment control automatically)
|
14218
|
Thu Sep 27 14:02:55 2018 |
yuki | Configuration | ASC | PZT driver board verification |
[ Yuki, Gautam ]
I fixed the input terminal that had been off, and made sure PZT driver board performs as we expect.
At first I ran a simulation of the PZT driver circuit using LTspice (Attached #1 and #2). It shows that when the bias is 30V the driver performs well only with high input volatage (bigger than 3V). Then I measured the performance as following way:
- Applied +-15V to the board with an expansion card and 31.8V to the high voltage port which is the maximum voltage of PS280 DC power supplier C10013.
- Terminated input and connectd input bias to GND, then set offset to -10.4V. This value is refered as elog:40m/8832.
- Injected DC signal into input port using a function generator.
- Measured voltage at the OUT port and MON port.
The result of this is attached #3 and #4. It is consistent with simulated one. All ports performed well.
- V(M1_PIT_OUT) = -4.86 *Vin +49.3 [V]
- V(M1_YAW_OUT) = -4.86 *Vin +49.2 [V]
- V(M2_PIT_OUT) = -4.85 *Vin +49.4 [V]
- V(M2_YAW_OUT) = -4.86 *Vin +49.1 [V]
- V(M1_PIT_MON) = -0.333 *Vin +3.40 [V]
- V(M1_YAW_MON) = -0.333 *Vin +3.40 [V]
- V(M2_PIT_MON) = -0.333 *Vin +3.40 [V]
- V(M2_YAW_MON) = -0.333 *Vin +3.40 [V]
The high voltage points (100V DC) remain to be tested. |
14219
|
Sun Sep 30 20:14:51 2018 |
yuki | Configuration | ASC | QPD calibration |
[ Yuki, Gautam, Steve ]
Results:
I calibrated a QPD (D1600079, V1009) and made sure it performes well. The calibration constants are as follows:
X-Axis: 584 mV/mm
Y-Axis: 588 mV/mm
Details:
The calibration of QPD is needed to calibrate steeing PZT mirrors. It was measured by moving QPD on a translation stage. The QPD was connected to its amplifier (D1700110-v1) and +-18V was supplied from DC power supplier. The amplifier has three output ports; Pitch, Yaw, and Sum. I did the calibration as follows:
- Center beam spot on QPD using steering mirror, which was confirmed by monitored Pitch and Yaw signals that were around zero.
- Kept Y-axis micrometer fixed, moved X-axis micrometer and measured the outputs.
- Repeated the procedure for the Y-axis.
The results are attached. The main signal was fitted with error function and I drawed a slope at zero crossing point, which is calibration factor. I determined the linear range of the QPD to be when the output was in range -50V to 50V, then corresponding displacement range is about 0.2 mm width. Using this result, the PZT mirrors will be calibrated in linear range of the QPD tomorrow.
Comments:
- Some X-Y coupling existed. When one axis micrometer was moved, a little signal of the other direction was also generated.
- As Gautam proposed in the previous study, there is some hysteresis. That process would bring some errors to this result.
- A scale of micrometer is expressed in INCH!
- The micrometer I used was made to have 1/2 inch range, but it didn't work well and the range of X-axis was much narrower.
Reference:
previous experiment by Gautam for X-arm: elog:40m/8873, elog:40m/8884 |
14220
|
Mon Oct 1 12:03:41 2018 |
not yuki | Configuration | ASC | PZT driver board verification |
I assume this QPD set is a D1600079/D1600273 combo.
How much was the SUM output during the measurement? Also how much were the beam radii of this beam (from the error func fittings)?
Then the calibration [V/m] is going to be the linear/inv-linear function of the incident power and the beam radus.
You mean the linear range is +/-50mV (for a given beam), I guess.
|
14221
|
Mon Oct 1 13:33:55 2018 |
yuki | Configuration | ASC | QPD calibration |
Quote: |
I assume this QPD set is a D1600079/D1600273 combo.
How much was the SUM output during the measurement? Also how much were the beam radii of this beam (from the error func fittings)?
Then the calibration [V/m] is going to be the linear/inv-linear function of the incident power and the beam radus.
You mean the linear range is +/-50mV (for a given beam), I guess.
|
- The SUM output was from -174 to -127 mV.
- The beam radii calculated from the error func fittings was 0.47 mm.
- Total optical path length measured by a ruler= 36 cm.
- Beam power measured at QPD was 2.96 mW. (There are some loss mechanism in the setup.)
Then the calibration factor of the QPD is
X axis: 584 * (POWER / 2.96mW) * (0.472mm / RADIUS) [mV/mm]
Y axis: 588 * (POWER / 2.96mW) * (0.472mm / RADIUS) [mV/mm]. |
14222
|
Mon Oct 1 20:39:09 2018 |
gautam | Configuration | ASC | c1asy |
We need to set up a copy of the c1asx model (which currently runs on c1iscex), to be named c1asy, on c1iscey for the green steering PZTs. The plan discussed at the meeting last Wednesday was to rename the existing model c1tst into c1asy, and recompile it with the relevant parts copied over from c1asx. However, I suspect this will create some problems related to the "dcuid" field in the CDS params block (I ran into this issue when I tried to use the dcuid for an old model which no longer exists, called c1imc, for the c1omc model).
From what I can gather, we should be able to circumvent this problem by deleting the .par file corresponding to the c1tst model living at /opt/rtcds/caltech/c1/target/gds/param/, and rename the model to c1asy, and recompile it. But I thought I should post this here checking if anyone knows of other potential conflicts that will need to be managed before I start poking around and breaking things. Alternatively, there are plenty of cores available on c1iscey, so we could just set up a fresh c1asy model...
|
- (write programming code of making alignment control automatically)
|
|
14224
|
Tue Oct 2 18:50:53 2018 |
yuki | Configuration | ASC | PZT mirror calibration |
[ Yuki, Gautam ]
I calibrated PZT mirrors. The ROUGH result was attached. (Note that some errors and trivial couplings coming from inclination of QPD were not considered here. This should be revised and posted again.)
The PZT mirrors I calibrated were:
- A 2-inch CVI mirror (45 degree, HR and AR for 532nm)
- A 1-inch Laseroptik mirror (45 degree, HR and AR for 532nm)
I did the calibration as follows:
- +-15V was supplied to PZT driver circuit, +100V to PZT driver bias, and +-18V to QPD amplifier.
- Optical path length was set to be same as that when I calibrated QPD, which is 36cm.
- The full range of CVI mirror is 3.5mrad according to its datasheet and linear range of QPD is 0.2mm, so I set the distance between PZT mirrors and QPD to be about 6cm. (I realized it was wrong. When mirror tilts 1 deg, the angle of beam changes 2 deg. So the distance should be the half.)
- After applying 0V to PZT driver input (at that time 50V was applied to PZT mirror), then centered beam spot on QPD using steering mirror, which was confirmed by monitored Pitch and Yaw signals of QPD that were around zero.
- In order to avoid hysteresis effect, I stared with an input signal of -10V. I then increased the input voltage in steps of 1V through the full range from -10V to +10V DC. The other input was kept 0V.
- Both the X and Y coordinates were noted in the plot in order to investigate pitch-yaw coupling.
The calibration factor was
CVI-pitch: 0.089 mrad/V
CVI-yaw: 0.096 mrad/V
Laseroptic-pitch: 0.062 mrad/V
Laseroptic-yaw: 0.070 mrad/V
Comments:
- I made sure that PZT mirrors move linearly in full input range (+-10V).
- PZT CH1 input: Yaw, CH2: Pitch, CH3: +100V bias
- The calibration factor of PZT mirrors [mrad/V] are not consistent with previous calibration (elog:40m/8967). I will check it again.
- I measured the beam power in order to calibrate QPD responce with a powermeter, but it didn't have high precision. So I used SUM output of QPD to the calibration.
- Full range of PZT mirrors looks 2 times smaller.
Reference:
Previous calibration of the same mirrors, elog:40/8967 |
14226
|
Wed Oct 3 14:24:40 2018 |
yuki | Configuration | ASC | Y end table upgrade plan |
Interim Procedure Report:
Purpose
The current setup of AUX Y-arm Green locking has to be improved because:
- current efficiency of mode matching is about 50%
- current setup doesn't separate the degrees of freedom of TEM01 with PZT mirrors (the difference of gouy phase between PZT mirrors should be around 90 deg)
- we want to remotely control PZT mirrors for alignment
What to do
- Design the new setup and order optices needed (finished!)
- As the new setup I designed, adding a new lens and slightly changing the position of optics are only needed. The new lens was arrived here.
- Check electronics (PZT, PZT driver, high voltage, cable, anti-imaging board) (finished!)
- All electronics were made sure performing well.
- The left thing to do is making a cable. (Today's tasks)
- Calibrate PZT mirror [mrad/V] (finished!)
- The result was posted here --> elog:40m/14224.
- Measure the status value of the current setup (power of transmitted light ...etc) (Tomorrow, --> finished!)
- Install them in the Y-end table and align the beam (Will start from Tomorrow) (The setup has a probrem I found on 10/04)
- Measure the status value of the new setup
- I want to finish above during my stay.
- Prepare the code of making alignment automaticaly
|
14227
|
Wed Oct 3 18:15:34 2018 |
yuki | Configuration | ASC | AI board improvement |
[ Yuki, Gautam ]
I improved Anti-Imaging board (D000186-Rev.D), which will be put between DAC port and PZT driver board.
It had notches at f = 16.6 kHz and 32.7 kHz, you can see them in the plot attached. So I replaced some resistors as follows:
- R6 and R7 replaced with 511 ohm (1206 thin film resistor)
- R8 replaced with 255 ohm (1206 thin film resistor)
- R14 and R15 replaced with 549 ohm (1206 thin film resistor)
- R16 replaced with 274 ohm (1206 thin film resistor)
Then the notch moved to 65.9 kHz (> sampling frequency of DAC = 64 kHz, good!).
(The plot enlarged around the notch frequency and the plot of all channels will be posted later.)
All electronics and optics seem to be ready.
Reference, elog:40m/8857
Diagram, D000186-D.pdf |
14228
|
Thu Oct 4 00:44:50 2018 |
yuki | Configuration | ASC | AI board improvement |
[ Yuki, Gautam ]
I made a cable which connects DAC port (40 pins) and AI board (25 pins). I will check if it works.
Tomorrow I will change setup for improvement of AUX Y-end green locking. Any optics for IR will not be moved in my design, so this work doesn't affect Y-arm locking with main beam.
While doing this work, I will do:
- check if the cable works
- make another cable which connects AI board (10 pins) and PZT driver (10 pins).
- check if eurocate in Y-rack (IY4?) applies +/-5V, +/-15V and +/-24V. It will be done using an expansion card.
- improve alignment servo for X-end.
- setup alignment servo for Y-end.
- about optical loss measurement.
|
14230
|
Thu Oct 4 22:15:30 2018 |
yuki | Configuration | ASC | Y-end table upgrade |
Before changing setup at Y-end table, I measured the status value of the former setup as follows. These values will be compared to those of upgraded setup.
- beam power going into doubling crystal (red12): 20.9 mW with filter, 1064nm
- beam power going out from doubling crystal (red12): 26.7 mW with filter, 532nm
- beam power going into faraday isolator (green5): 0.58 mW without filter, 532nm
- beam power going out from faraday isolator (green5): 0.54 mW without filter, 532nm
- beam power going to ETMY: 0.37 mW without filter, 532nm
- beam power of transmitted green light of Y-arm, which was measured by C1:ALS-TRY_OUT: 0.5 (see attachment #1)
(These numbers are shown in the attachment #2.)
The setup I designed is here. It can bring 100% mode-matching and good separation of degrees of TEM01, however I found a probrem. The picture of setup is attached #3. You can see the reflection angle at Y7 and Y8 is not appropriate. I will consider the schematic again. |
14231
|
Fri Oct 5 00:46:17 2018 |
Koji | Configuration | ASC | Y-end table upgrade |
???
The SHG crystal has the conversion efficiency of ~2%W (i.e. if you have 1W input @1064, you get 2% conversion efficiency ->20mW@532nm)
It is not possible to produce 0.58mW@532nm from 20.9mW@1064nm because this is already 2.8% efficiency.
|
14232
|
Fri Oct 5 17:32:38 2018 |
yuki | Configuration | ASC | Y-end table upgrade |
I measured it with the wrong setting of a powermeter. The correct ones are here:
- beam power going into doubling crystal (red12): 240 mW, 1064nm
- beam power transmitted dichroic mirror (Y5): 0.70 mW, 532nm
- beam power going into faraday isolator (green5): 0.58 mW, 532nm
- beam power going out from faraday isolator (green5): 0.54 mW, 532nm
- beam power going to ETMY: 0.37 mW, 532nm
- beam power of transmitted green light of Y-arm, which was measured by C1:ALS-TRY_OUT: 0.5 (see attachment #1)
The calculated conversion efficiency of SHG crystal is 1.2%W. |
14233
|
Fri Oct 5 17:47:55 2018 |
gautam | Configuration | ASC | Y-end table upgrade |
What about just copying the Xend layout? I think it has good MM (per calculations), reasonable (in)sensitivity to component positions, good Gouy phase separation, and I think it is good to have the same layout at both ends. Since the green waist has the same size and location in the doubling crystal, it should be possible to adapt the X end solution to the Yend table pretty easily I think.
Quote: |
The setup I designed is here. It can bring 100% mode-matching and good separation of degrees of TEM01, however I found a probrem. The picture of setup is attached #3. You can see the reflection angle at Y7 and Y8 is not appropriate. I will consider the schematic again.
|
|
14234
|
Fri Oct 5 22:49:22 2018 |
yuki | Configuration | ASC | Y-end table upgrade |
I designed a new layout. It has good mode-matching efficiency, reasonable sensitivity to component positions, good Gouy phase separation. I'm setting optics in the Y-end table. The layout will be optimized again after finishing (rough) installation. (The picture will be posted later) |
14235
|
Sun Oct 7 16:51:03 2018 |
gautam | Configuration | LSC | Yarm triggering changed |
To facilitate Yuki's alignment of the EY green beam into the Yarm cavity, I have changed the LSC triggering and PowNorm settings to use only the reflected light from the cavity to do the locking of Arm Cavity length to PSL. Running the configure script should restore the usual TRY triggering settings. Also, the X arm optics were macroscopically misaligned in order to be able to lock in this configuration. |
14236
|
Sun Oct 7 22:30:42 2018 |
yuki | Configuration | LSC | Yarm Green locking was recovered |
I finished installation of optics in the Y-end and recovered green locking. Current ALS-TRY_OUTPUT is about 0.25, which is lower than before. So I still continue the alignment of the beam. The simulation code was attached. (Sorry. The optic shown as QWP2 is NOT QWP. It's HWP.) |
14237
|
Mon Oct 8 00:46:35 2018 |
yuki | Configuration | ASC | Y-end table upgrade |
Quote: |
I measured it with the wrong setting of a powermeter. The correct ones are here:
- beam power going into doubling crystal (red12): 240 mW, 1064nm
- beam power transmitted dichroic mirror (Y5): 0.70 mW, 532nm
- beam power going into faraday isolator (green5): 0.58 mW, 532nm
- beam power going out from faraday isolator (green5): 0.54 mW, 532nm
- beam power going to ETMY: 0.37 mW, 532nm
- beam power of transmitted green light of Y-arm, which was measured by C1:ALS-TRY_OUT: 0.5 (see attachment #1)
|
After installation I measured these power again.
- beam power going into doubling crystal: 241 mW, 1064nm
- beam power transmitted dichroic mirror: 0.70 mW, 532nm
- beam power going into faraday isolator: 0.56 mW, 532nm
- beam power going out from faraday isolator: 0.53 mW, 532nm
- beam power going to ETMY: 0.36 mW, 532nm
There is a little power loss. That may be because of adding one lens in the beam path. I think it is allowable margin. |
14238
|
Mon Oct 8 18:56:52 2018 |
gautam | Configuration | ASC | c1asx filter coefficient file missing |
While pointing Yuki to the c1asx servo system, I noticed that the filter file for c1asx is missing in the usual chans directory. Why? Backups for it exist in the filter_archive subdirectory. But there is no current file. Clearly this doesn't seems to affect the realtime code execution as the ASX model seems to run just fine. I copied the latest backup version from the archive area into the chans directory for now. |
14239
|
Tue Oct 9 16:05:29 2018 |
gautam | Configuration | ASC | c1tst deleted, c1asy deployed. |
Setting up c1asy:
- Backed up old c1tst.mdl as c1tst_old_bak.mdl in /opt/rtcds/userapps/release/cds/c1/models
- Copied the c1tst model to /opt/rtcds/userapps/release/isc/c1/models/c1asy.mdl as this is where the c1asx.mdl file resides.
- Backed up original c1rfm.mdl as c1rfm_old.mdl in /opt/rtcds/userapps/release/cds/c1/models (since the old c1tst had an RFM block which is unnecessary).
- Deleted offending RFM block from c1rfm.mdl.
- Recompiled and re-installed c1rfm.mdl. Model has not yet been restarted, as I'd like suspension watchdogs to be shutdown, but c1susaux EPICS channels are presently not responsive.
- Removed c1tst model (C-node91) from /opt/rtcds/caltech/c1/target/gds/param/testpoints.
- Removed /opt/rtcds/caltech/c1/target/gds/param/tpchn_c1tst.par (at this point, DCUID 91 is free for use by c1asy).
- Moved c1tst line in /opt/rtcds/caltech/c1/target/daqd/master to "old model definitions models" section.
- Added /opt/rtcds/caltech/c1/target/gds/param/tpchn_c1asy.par to the master file.
- Edited/diskless/root.jessie/etc/rtsystab to allow c1asy to be run on c1iscey.
- Finally, I followed the instructions here to get the channels into frames and make all the indicators green.
Now Yuki can work on copying the simulink model (copy c1asx structure) and implementing the autoalignment servo. |
14240
|
Tue Oct 9 23:03:43 2018 |
yuki | Configuration | LSC | Yarm Green locking was recovered |
[ Yuki, Gautam, Steve ]
To align the green beam in Y-end these hardware were installed:
- PZT mirrors in Y-end table
- PZT driver in 1Y4 rack
- Anti-Imaging board in 1Y4 rack
- cables (DAC - AIboard - PZTdriver - PZT)
- high voltage supplier
I made sure that DAC CH9~16 and cable to AI-board worked correctly.
When we applied +100V to PZT driver and connected DAC, AI-board and PZT drive, the output voltage of the driver was not correct. I'll check it tomorrow. |
14241
|
Wed Oct 10 12:38:27 2018 |
yuki | Configuration | LSC | All hardware was installed |
I connected DAC - AIboard - PZTdriver - PZT mirrors and made sure the PZT mirrors were moving when changing the signal from DAC. Tomorrow I will prepare alignment servo with green beam for Y-arm. |
14257
|
Mon Oct 15 20:11:56 2018 |
yuki | Configuration | ASC | Y end table upgrade plan |
Final Procedure Report for Green Locking in YARM:
Purpose
The current setup of AUX Y-arm Green locking has to be improved because:
- current efficiency of mode matching is about 50%
- current setup doesn't separate the degrees of freedom of TEM01 with PZT mirrors (the difference of gouy phase between PZT mirrors should be around 90 deg)
- we want to remotely control PZT mirrors for alignment
What to do
- Design the new setup and order optices needed (finished!)
- As the new setup I designed, adding a new lens and slightly changing the position of optics are only needed. The new lens was arrived here.
- Check electronics (PZT, PZT driver, high voltage, cable, anti-imaging board) (finished!)
- All electronics were made sure performing well.
- The left thing to do is making a cable. (Today's tasks)
- Calibrate PZT mirror [mrad/V] (finished!)
- The result was posted here --> elog:40m/14224.
- Measure the status value of the current setup (power of transmitted light ...etc) (finished!)
- Install them in the Y-end table and align the beam (Almost finished!) (GTRY signal is 0.3 which means Mode-Matching efficiency is about 30%. It should be improved.)
- Measure the status value of the new setup (finished!)
- Prepare the code of making alignment automaticaly
- see sitemap.adl>ASC>c1asy. I prepared medm. If you move PZT SLIDERS then you can see the green beam also moves.
- Preparing filters is needed. You can copy them from C1ASX.
- Note that now you cannot use C1ASX servo because filters are not applied.
|
14260
|
Wed Oct 17 20:46:24 2018 |
yuki | Configuration | ASC | Y end table upgrade plan |
To do for Green Locking in YARM:
The auto-alignment servo should be completed. This servo requires many parameters to be optimized: demodulation frequency, demodulation phase, servo gain (for each M1/2 PIT/YAW), and matrix elements which can remove PIT-YAW coupling. |
14301
|
Fri Nov 16 15:09:31 2018 |
Steve | Configuration | VAC | not venting cryo and ion pumps |
Notes on the ion pumps and cryo pump:
-
Our 4 ion pumps were closed off for a lomg time. I estmated their pressure to be around ~1 Torr. After talking with Koji we decided not to vent them.
-
It'd be still useful to wire their position sensors. But make sure we do not actuate the valves.
-
The cryo pump was regenerated to 1e-4 Torr about 2 years ago. It's pressure can be ~ 2 Torr with charcoal powder. It is a dirty system at room temperature.
-
Do not actuate VC1 and VC2, and keep its manual valve closed.
-
IF someone feels we should vent them for some reason, let us know here in the elog before Monday morning.
Quote: |
Wiring of the power, Ethernet, and indicator lights for the vacuum Acromag chassis is complete. Even though this crate will only use +24V DC, I wired the +/-15V connector and indicator lights as well to conform to the LIGO standard. There was no wiring diagram available, so I had to reverse-engineer the wiring from the partially complete c1susaux crate. Attached is a diagram for future use. The crate is ready to begin software developing on Monday.
|
|
14322
|
Tue Nov 27 17:06:51 2018 |
Steve | Configuration | VAC | Agilent 84FS turbo installed as TP2 |
Chub & Steve,
We swapped in our replacement of Varian V70D "bear-can" turbo as factory clean.
The new Agilent TwisTorr 84 FS turbo pump [ model x3502-64002, sn IT17346059 ] with intake screen, fan, vent valve. The controller [ model 3508-64001, sn IT1737C383 ] and a larger drypump IDP-7, [ model x3807-64010, sn MY17170019 ] was installed.
Next things to do:
- implement hardware interlock to close V4 at 80% pumping speed slowdown of "standby" rotation speed, estimated to be ~ 40,000 RPM ( when Standby 50K RPM )
- set up isolation valve in the foreline of TP2, with delayed start of the IDP-7 and/or use relay to power drypump. This turbo controller can not switch off or start of the dry pump. [ Agilent isolation valve #X3202-60055, with position indicator, pneumatic actuation, 115V solenoid ]..........as a second thought, we do not need isolation valve if we go with the relay option. The IDP-7 has built in delay of 10-15 sec
- test performance of new turbo
|
14382
|
Thu Jan 3 21:17:49 2019 |
rana | Configuration | Computers | Workstation Upgrade: Donatella -> Scientific Linux 7.2 |
donatella was one of our last workstations running ubuntu12. we installed SL7 on there today
- had to use a DVD; wouldn't boot from USB stick
- made sure to use userID=1001 and groupID=1001 at the initial install part
- went to the Keith Thorne LLO wiki on SL7
- The 'yum update' command failed due to a gstreamer conflict. I did "yum remove gstreamer1-plugins-ugly-free-1.10.4-3.el7.x86_64" and then it continued a bit more.
-
Then there are ~20 errors related to gds-crtools that look like this:Error: Package: gds-crtools-2.18.12-1.el7.x86_64 (lscsoft-production) Requires: libMatrix.so.6.14()(64bit)
- I re-ran the yum install .... command using the --skip-broken command and that seemed to complete, although I guess the GDS stuff will not work.
- Installed: terminator, inconsolata-fonts,
- Installed XFCE desktop as per K Thorne: yum groupinstall "Xfce" -y
-
|
14387
|
Mon Jan 7 11:54:12 2019 |
Jon | Configuration | Computer Scripts / Programs | Vac system shutdown |
I'm making a controlled shutdown of the vac controls to add new ADC channels. Will advise when it's back up. |
14388
|
Mon Jan 7 19:21:45 2019 |
Jon | Configuration | Computer Scripts / Programs | Vac system shutdown |
ADC work finished for the day. The vac controls are back up, with all valves CLOSED and all pumps OFF.
Quote: |
I'm making a controlled shutdown of the vac controls to add new ADC channels. Will advise when it's back up.
|
|
14476
|
Fri Mar 8 08:40:26 2019 |
Anjali | Configuration | | Frequency stabilization of 1 micron source |
The schematic of the homodyne configuration is shown below.
Following are the list of components
Item |
Quantity |
Availability |
Part number |
Remarks |
Laser (NPRO) |
1 |
Yes |
|
|
Couplers (50/50) |
5 |
3 No's |
FOSC-2-64-50-L-1-H64F-2 |
Fiber type : Hi1060 Flex fiber |
Delay fiber |
two loops of 80 m |
Yes |
PM 980 |
One set of fiber is now kept along the arm of the interferometer
|
InGaAs PD (BW > 100 MHz) |
4 |
Yes |
NF1611 |
Fiber coupled (3 No's)
Free space ( 2 No's)
|
SR560 |
3 |
Yes |
|
|
- The fiber mismatch between the couplers and the delay fiber could affect the coupling efficiency
|
14664
|
Tue Jun 11 19:25:58 2019 |
aaron | Configuration | BHD | Reviving the single OMC BHD design? |
I drew out some idea of how we might use a single OMC to clean both paths of the BHD after mixing, without being susceptible to polarization-dependent effects within the OMC. Basically, can we send the two legs of the BHD into the OMC counterpropagating. I've attached a diagram.
I think one issue would be scattered light, since any backscatter directly couples into the counterpropagating mode, and thus directly to the PD. However, unless the polarization of the scattered light rotates it would not scatter back to the IFO. And, since the LO and signal mix before the OMC, this scattered light would not directly add phase noise.
Maybe more problematic would be that if the rejection at the PBS (or the polarization rotation) isn't perfect, light from the LO directly couples into the dark port. Can we get away with a Faraday isolator before the OMC?
Diagram attached. |
14672
|
Thu Jun 13 22:21:44 2019 |
Koji | Configuration | CDS | Paola wireless connected to martian |
SURFs had trouble connecting paola to martian via wireless.
Of course, it requires a fixed IP but it had not it yet. So I went to chiara and gave 192.168.113.110 as "paolawl". Note that the wired connection has .111 and it is "paola".
Followed the instruction on http://nodus.ligo.caltech.edu:8080/40m/14121 |
14685
|
Fri Jun 21 19:22:40 2019 |
Koji | Configuration | BHD | Reviving the single OMC BHD design? |
I think a Faraday rotator rotates the polarizations in a same way for both forward and backward beam, and it's not like in this figure.
And the transmission through multiple faradays will also be a big issue. |
14692
|
Mon Jun 24 13:48:36 2019 |
Kruthi | Configuration | CDS | Giada wireless connection |
[Gautam, Kruthi]
This afternoon, Gautam helped me setup Giada to access the GigE installed for MC2. Unlike Paola, which was being used earlier, Giada has a better battery life and doesn't shutdown when the charger is unplugged. Gautam configured Giada to enable its wireless connection to Martian, just like Koji had configured Paola (https://nodus.ligo.caltech.edu:8081/40m/14672). We also rerouted the ethernet cable we were using with the PoE adaptor from Netgear Switch in 1x2 to 1x6. |
14767
|
Wed Jul 17 17:56:18 2019 |
Koji | Configuration | Computers | Gave resolv.conf to giada |
Kruthi noticed that she could not login to rossa from giada.
I checked /etc/resolv.conf and it was
nameserver 127.0.0.1
so obviously it is useless to refer localhost (i.e. giada) as a nameserver.
I copied our usual resolv.conf to giada as following:
nameserver 192.168.113.104
nameserver 131.215.125.1
nameserver 8.8.8.8
search martian
Giada's ssh known_host had unupdated entry for rossa, so I had to clean it up, but after that we can connect to rossa from giada just by "ssh rossa".
Case closed. |
14812
|
Thu Jul 25 14:28:03 2019 |
gautam | Configuration | Computers | firewalld disabled for EPICS CA |
I think rana did some more changes to this workstation to make it useful for commissioning activities - but the MEDM screens were still white blanks. The problem was that the firewalld wasn't disabled (last two steps of the KThorne setup wiki). I disabled it. Now donatella can run MEDM, ndscope and StripTool. DTT doesn't work to get online data because of a "Synchronization Error", I'm not bothering with this for now. I think Kruthi successfully demonstrated the fetching of offline data with DTT. |
15085
|
Sun Dec 8 20:48:29 2019 |
rana | Configuration | Computers | Megatron: starts up grade |
I noticed recently that Megatron was running Ubuntu 12, so I've started its OS upgrade.
- Unlocked the IMC + disabled the autolocker from the LockMC screen + closed the PSL shutter (IMC REFL shutter doesn't seem to do anythin)
- Disabled the "FSS" slow servo on the FSS screen
- did sudo apt-get update, sudo apt-get upgrade, and then sudo apt-get do-release-upgrade which starts the actual thing
- According to the internet, the LTS upgrades will go in series rather than up to 18 in one shot, so its now doing 12 -> 14 (Trusty Tapir)
Megatron and IMC autolocking will be down for awhile, so we should use a different 'script' computer this week.
Mon Dec 9 14:52:58 2019
upgrade to Ubuntu 14 complete; now upgrading to 16 |
15095
|
Wed Dec 11 22:01:24 2019 |
rana | Configuration | Computers | Megatron: starts up grade |
Megatron is now running Ubuntu 18.04 LTS.
We should probably be able to load all the LSC software on there by adding the appropriate Debian repos.
I have re-enabled the cron jobs in the crontab.
The MC Autolocker and the PSL NPRO Slow/Temperature control are run using 'initctl', so I'll leave that up to Shruti to run/test. |
15117
|
Mon Jan 13 15:47:37 2020 |
shruti | Configuration | Computer Scripts / Programs | c1psl burt restore |
[Yehonathan, Jon, Shruti]
Since the PMC would not lock, we initially burt-restored the c1psl machine to the last available shapshot (Dec 10th 2019), but it still would not lock.
Then, it was burt-restored to midnight of Dec 1st, 2019, after which it could be locked. |
15125
|
Wed Jan 15 14:10:28 2020 |
Jon | Configuration | PSL | New EPICS database for C1PSL + C1IOO |
Summary
I have completed the new EPICS channel database for the c1psl and c1ioo channels (now combined into the new c1psl Acromag machine). I've tested a small subset of channels on the electronics bench to confirm that the addressing and analog channel calibrations are correct in a general sense. At this point, we are handing the chassis off to Chub to complete the wiring of the Acromag terminals to Dsub feedthroughs. At the 40m meeting today, we identified Feb. 17-22 as a potential window for installation in the interferometer (Gautam is out of town then). Below are some implementaton details for future reference.
Analog channel calibration for Acromag
For analog input (ai) channels, the Acromag outputs raw values ranging from +/-30,000 counts, but the EPICS IOC interprets the data type as ranging from +/-2^15 = 32,768. Similarly, for analog output (ao) channels, the Acromag expects a drive signal in the range +/-30,000 counts. To achieve proper scaling, Johannes had previously changed the EGUF and EGUL fields from +/-10 V to +/-10.923 V. However, changing the engineering fields makes it much harder for a human to read off the real physical I/O range of the channel.
A better way to achieve the correct scaling is to simply set the field ASLO=1.09225 (65,536 / 60,001) in addition to the normal EGUF and EGUL field values (+/-10 V). Setting this field forces a rescaling of the number of raw counts that works as so (assuming a 16-bit bipolar ADC or DAC, as are the Acromags):
OVAL = (RVAL * ASLO + AOFF + 2^15) * (EGUF - EGUL) / 2^16 + EGUL
In the above mapping, OVAL is the value of the channel in engineering units (e.g., V) and RVAL is its raw value in counts. It is not the case that either the ASLO/AOFF or EGUF/EGUL fields are used, but not both. The ASLO/AOFF parameters are always applied (but their default values are ASLO=1 and AOFF=0, so have no effect unless changed). The EGUF and EGUL parameters are then additionally applied if the field LINR="LINEAR" is set.
This conversion allows the engineering fields to remain unchanged from the real physical range. The ASLO value is the same for both analog input and output channels. I have implemented this on all the new c1psl and c1ioo channels and confirmed it to work using a calibrated input voltage source. |
15142
|
Wed Jan 22 19:17:20 2020 |
gautam | Configuration | Computers | Megatron: starts up grade |
upgrade was done
cronjob testing wasn't one by one 😢
burt snapshots were gone
i brought them back home 🏠
Quote: |
Megatron is now running Ubuntu 18.04 LTS.
|
|
15145
|
Thu Jan 23 15:32:42 2020 |
gautam | Configuration | Computers | Megatron: starts up grade |
The burt snapshotting is still not so reliable - for whatever reason, the number of snapshot files that actually get written looks random. For example, the 14:19 backup today got all the snaps, but 15:19 did not. There are no obvious red flags in either the cron job logs or the autoburt log files. I also don't see any clues when I run the script in a shell. It'll be good if someone can take a look at this. |
15150
|
Thu Jan 23 23:07:04 2020 |
Jon | Configuration | PSL | c1psl breakout board wiring |
To facilitate wiring the c1psl chassis and scripting loopback tests, I've compiled a distilled spreadsheet with the Acromag-to-breakout board wiring, broken down by connector. This information is extractable from the master spreadsheet, but not easily. There were also a few apparent typos which are fixed here.
The wiring assignments at the time of writing are attached below. Here is the link to the latest spreadsheet. |
15158
|
Mon Jan 27 14:01:01 2020 |
Jordan | Configuration | General | Repurposed Sorenson Power Supply |
The 24 V Sorenson (2nd from bottom) in the small rack west of 1x2 was repurposed to 12V 600 mA, and was run to a terminal block on the north side of 1X1. Cables were routed underneath 1X1 and 1X2 to the terminal blocks. 12V was then routed to the PSL table and banana clip terminals were added. |
15159
|
Mon Jan 27 18:16:30 2020 |
gautam | Configuration | Computers | Sluggish megatron? |
I've also been noticing that the IMC Autolocker scripts are running rather sluggishly on Megatron recently. Some evidence - on Feb 11 2019, the time between the mcup script starting and finishing is ~10 seconds (I don't post the raw log output here to keep the elog short). However, post upgrade, the mean time is more like ~45-50 seconds. Rana mentioned he didn't install any of the modern LIGO software tools post upgrade, so maybe we are using some ancient EPICS binaries. I suspect the cron job for the burt snapshot is also just timing out due to the high latency in channel access. Rana is doing the software install on the new rossa, and once he verifies things are working, we will try implementing the same solution on megatron. The machine is an old Sun Microsystems one, but the system diagnostics don't signal any CPU timeouts or memory overflows, so I'm thinking the problem is software related...
Quote: |
The burt snapshotting is still not so reliable - for whatever reason, the number of snapshot files that actually get written looks random. For example, the 14:19 backup today got all the snaps, but 15:19 did not. There are no obvious red flags in either the cron job logs or the autoburt log files. I also don't see any clues when I run the script in a shell. It'll be good if someone can take a look at this.
|
|
15164
|
Tue Jan 28 15:39:04 2020 |
gautam | Configuration | Computers | Sluggish megatron? |
There were a bunch of medm processes stalled on megatron (connected with screenshot taking). To see if they were interfering with the other scripts, I killed all of the medm processes, and commented out the line in the crontab that runs the screenshots every 10 mins. Let's see if this improves stability. |
15167
|
Tue Jan 28 17:36:45 2020 |
gautam | Configuration | Computers | Local EPICS7.0 installed on megatron |
[Jon, gautam]
We found that the caput commands were taking much longer to execute on megatron than on pianosa (for example). Suspecting that this had something to do with the fact that megatron was using EPICS binaries from the shared NFS drive which were compiled for a much older OS, I installed the latest stable release of EPICS on megatron. The new caput commands execute much faster. I also added the local EPICS directory to the head of the $PATH variable used by the MC autolocker and FSS Slow scripts, so that they use the new caput command. But mcup is still slow - maybe my new path definition isn't picked up and it is still using the NFS binaries? To be looked into...
Quote: |
There were a bunch of medm processes stalled on megatron (connected with screenshot taking). To see if they were interfering with the other scripts, I killed all of the medm processes, and commented out the line in the crontab that runs the screenshots every 10 mins. Let's see if this improves stability.
|
|