5666
|
Fri Oct 14 16:20:11 2011 |
Zach | Update | SUS | C1:SUS-ETMX_SPDMon fixed |
I offered to help Kiwamu with some of the 40m work. The first task was to figure out why the ETMX side OSEM monitor was so low, since we know that the depth is about right. It was showing ~0.13 V to the others' ~0.7 V.
TL,WR: There was a wire disconnected from the breakout panel on the side of the rack
I started by pulling the board out and checking to make sure that it was working properly. I injected a sine wave to the SIDE IN and found that it showed up in the signal coming out of the back (into the crate) just fine (see below). One strange thing I noticed while testing the board is that both inputs for each used channel of the MAX333 switches on the board are shorted to their respective outputs. That is, the switches seem to be open to BOTH 0 and 1 logic states. This seems counterintuitive, but perhaps there's something about how these work that I don't know.

Then I went about tracing the signal from the back of the crate to the breakout panel on the side of the rack. I opened it up, verified that the ribbon cable was transmitting correctly, and as I went to plug it back in I noticed that one of the wires---the correct one---had come completely undone.

The screw clamp appeared to be a bit finicky, as I had to loosen and tighten it a few times before it finally seemed to grab hold of the wire. It probably just wasn't tight in the first place and the wire was pulled out. Anyway, things are working now:

|
5665
|
Fri Oct 14 04:35:45 2011 |
kiwamu | Update | LSC | locking tonight |
The lock of DRMI wasn't stable enough to measure the sensing matrix. Failed.    
PRMI and SRMI were okay and in fact they could stay locked robustly for a long time.
I added a new option in the C1IFO_CONFIGURE screen so that one can choose Signal-Recycled Michelson in carrier resonant condition.

Additionally the orthogonalization of the I-Q signals on REFL55 should be done because it hasn't been done.
|
5664
|
Thu Oct 13 23:58:38 2011 |
Koji | Update | LSC | fixing REFL165 |
I already have reported in this entry that REFL165 shows too high DC output which does not depend on the light level on the diode.
Today I removed REFL165 from the table and inspected it.
The diode has been burnt as shown in the first picture (left).
The window is smoked, and the photo sensitive surface has been removed from its base. It moves in the can.
The burnt diode was replaced to the new one.
The new one shows ~30% better capacitance of ~50pF and I had to increase the inductance from 14nH (i.e. 15nH//220nH) to 18nH.
After some struggles to increase/decrease the stray inductance by moving the SMD capacitors a little, the resonance is reasonably tuned to 166MHz.
The comprehensive test will be performed shortly. |
5663
|
Thu Oct 13 21:44:48 2011 |
Mirko | Update | CDS | Seismic BLRMS channels, new RMS calculation |
[Rana, Koji, Mirko]
We looked into the CDS RMS block c-code as described in Rolfs RCG app guide. Seems the block uses a first order LP filter with a corner freq. / time of 20k execution cycles. There are also some weird thersholds at +-2000counts in there.
I was looking into implementing a hand-made RMS block, by squaring, filtering, rooting. The new RMS (left) seems nicer than the old one (bottom right). Signal was 141counts sinus at 4Hz.
Filters used: Before squaring: 4th order butterworth BP at given freq. & (new) 6th order inverse Chebyshew 20dB at 0.9*lower BP freq. and 1.1*upper BP freq. => about 1dB at BP freq.
After squaring: 4th order butterworth LP @ 1Hz.
C1PEM execution time increased from about 20us to about 45us.
Made a new medm screen with the respective filters in place of the empty C1PEM_OVERVIEW. Should go onto the sitemap.

Original RMS LP is slower than 0.1Hz, see below for single LP at 0.1Hz in the new RMS. Original RMS is faster than single LP @ 0.01Hz

Some of the channels are recorded as 256Hz DAQ channels now. Need to figure out how to record these as 16Hz EPICS channls. |
5662
|
Thu Oct 13 21:40:59 2011 |
rana | Summary | VAC | Recovery from the power shutdown is completed |
As it turns out Steve is not crazy in this particular instance: the vacuum computer, linux3 , has some issues. I can login just fine, but trying to open a terminal makes the CPU rail and the RAM max out and eventually the machine freezes.
Under KDE, I can open a terminal OK as root, but if I then try a 'su controls', the same issue happens. Let's wait for Jamie to fix this. |
5661
|
Thu Oct 13 20:25:32 2011 |
Katrin | Update | Green Locking | LPF transfer function YARM |
It is a 4th order filter with cut of frequency of 120 kHz.
Design

Measurement

  

|
5660
|
Thu Oct 13 14:23:09 2011 |
steve | Update | SUS | ITMX oplev with 3 mm beam on qpd |
I replaced the JDSU-Uniphase 1125P by 1103P He/Ne laser. This new laser had 2.8 mW output yesterday. It degraded to 0.5 mW by this morning.
The beam size on the QPD is ~3 mm This should give us better sensitivity. These are not the perfect lenses at all, but we have them here.
On the other hand, there are still some coherence below 1 Hz, so the laser intensity noise or clipping dominating this part of the spectrum.
|
5659
|
Thu Oct 13 03:22:53 2011 |
kiwamu | Update | LSC | measurement of sensing matrix : just began |
- status update on LSC activity :
The measurement of the LSC sensing matrix has begun. But no useful results yet.
The measurement script (#4850) ran pretty well after I did some modifications to adopt the script to the latest LSC model.
However the SNR weren't so great particularly in REFL33 in the PRMI configuration.
So I will tune the amplitude of excitations and integration times tomorrow.
Currently the excitation is at 238.1 Hz, where no disturbing structures are found in the spectra. |
5658
|
Wed Oct 12 19:58:32 2011 |
Katrin | Update | Green Locking | Power splitter is unbalanced |
The mini circuit power splitter ZFRSC-42S+ used at the YARM has no balanced output as it should have according to the data sheet.
@ 0.05MHz the amplitude unbalance should be 0.03 dB
A quick measurement shows that there is a LO amplitude dependent unbalance:
LO amplitude input (Vpp) |
unbalanced output (dB) |
1.3 |
3.66 |
1.4 |
4.08 |
1.5 |
4.28 |
1.6 |
4.36 |
So my question is, shall I replace the power splitter just in case it is further degrading? |
5657
|
Wed Oct 12 18:54:02 2011 |
Katrin | Update | Green Locking | 60 Hz oscillation due to broken BNC cable |
There was a 60 Hz and 120 Hz oscillation on the green PDH photo diode output. After a long search, I could identify that
the source was a broken BNC cable which was connected to the photo diode. I exchanged that BNC cable and the 60 Hz
and 120 Hz are gone :-)
With the new cable the PD output was less noisy so that it was easier to achieve a better alignment of the light to the cavity.
The reflected power could be reduced from 40% to 30%. For perfect alignment the reflected power would be 20%. |
5656
|
Wed Oct 12 17:53:01 2011 |
kiwamu | Update | LSC | BS actuator response : fitting done and histroy of delays |
An update on calibration of the BS actuator : A fitting has been done.
(Fitting)
I used LISO for fitting the complex transfer function.
Because the data points around 1 Hz didn't have big coherence a few data points, which had coherence of less than 0.9, were excluded.
Also the fitting of the Q-factor wasn't successful due to the lack of good data points around the resonance.So I left Q fixed to be 5 in the fitting.
(Fitting result)
G = 2.18060874008e-8 +/- 6.425e-10 (2.95%)
f0 = 1.0100491195 +/- 1.51e-2 (1.49%) [Hz]
Q = 5 (fixed)
delay = 423.2753462089e-6 +/- 4.989e-6 (1.18%) [sec]

(History of delay)
Because we have been observing several different amount of delays in different configurations, perhaps it is worth to summarize those numbers.
description |
delay [usec] |
elog entry |
MICH lock (BS actuation) |
423 |
this entry |
LSC feed forward path |
127 |
#5218 |
MICH lock (BS actuation) |
600 |
#4638 |
ALS on X arm (ETMX actuation) |
330 |
#4196 |
RFM (from c1lsc to c1sus) |
125 |
#4153 |
from ADC to DAC (all the front end machine) |
38-110 |
#3961 |
from ADC to DAC (c1sus) |
124 |
#3838 |
RFM (c1ioo and c1sus) |
8-62 |
#3855 |
Quote from #5648 |
Tomorrow I will do a serious fitting.
|
|
5655
|
Wed Oct 12 08:43:30 2011 |
steve | Update | SUS | ITMX oplev improved a bit |
Atm2 is before optical path adjustment. The idea was to remove possible clipping in vacuum.
Coherense significantly reduced below 4 Hz
Today I will replace the He/Ne laser 1125P with 1103P
|
5654
|
Wed Oct 12 00:35:42 2011 |
kiwamu | Update | LSC | TRY path realigned |
The TRY (TRansmitted light from Y arm ) path was a bit realigned because there had been a small clipping.
This clipping was introducing offsets on the error signals of the C1ASS servo.
(Story)
During I was running the C1ASS servo on the Y arm I found every time after the auto-alignment is done there still remained a slight offset in the beam pointing,
I looked at the CCD camera which looks at the transmitted light and then introduced an intentional misalignment in ETMY in order to find an obvious clipping.
Indeed there was a clipping in horizontal direction. I checked through the optics on the Y end optical bench.
On the second mirror (beam splitter) the beam was on a very edge. So I steered the first steering mirror to fix it,
In addition to that an iris which is placed between the first and second mirror was also clipping the beam,
So I fully opened the aperture of the iris. |
5653
|
Tue Oct 11 21:23:51 2011 |
Jenne | Update | LSC | Arm absl lengths |
Quote: |
[Katrin, Jenne]
We took the data for the new absolute length measurement of both arms, after the latest vent and move. We will analyze soonly. We had done a round of analysis, but then Koji pointed out that our data wasn't so clean because the whitening filters were on (and saturated the ADC). We now have the data (but not the analysis) for the better data with the WF off.
So our dirty-data preliminary number for the X arm is 37.73meters, which is 14cm different from our old length. We were supposed to move by ~20cm, so....either this measurement is bad because the data sucked (which it did), or we are 6cm off. Or both.
I'll do another analysis with the clean data for both arms later today/tomorrow.
|
After analyzing the cleaner data, I get the following:
Y_Length_long = 37.757 meters
X_Length_long = 37.772 meters
As stated in the wiki, the goal arm length was L = 37.7974 m for each arm.
So we're within 2cm for X, and within 4cm for Y.
According to Kiwamu's awesome tolerance calculation, we need to be within 2cm for each arm. Given that we started out 20cm wrong for X and 25cm wrong for Y, we're a lot closer now, even though we aren't meeting our Yarm requirement yet.
Probably some Optickle action is in order, to see what these new lengths give us in terms of sideband phase and other stuff.
If you want more digits on my calculated numbers (which are probably meaningless, but I haven't done a careful error analysis), in my directory ...../users/jenne/Xarm and ..../users/jenne/Yarm run Xarm_find_peaks_and_length.m and Yarm_find_peaks_and_length.m respectively. These will output the lengths. |
5652
|
Tue Oct 11 19:11:25 2011 |
kiwamu | Update | LSC | Re: BS actuator reponse at low frequency : measured |
I think the precision due to the loop gain uncertainty is something like 0.1% at 0.1 Hz. It's not the issue.
The real issue was the loud motion of MICH, which degrades the coherence of the measurement.
Also last night I tried the fringe hopping technique and gave it up for several reasons.
(uncertainty due to the loop gain)
When MICH is locked, the signal at C1:LSC-MICH_OUT can be expressed in frequency domain by
MICH_OUT = G / (1+G) * (1 / A) * X + G / (1+G) * (1 / H) * (1 / A) * S, [1]
where G is the open loop gain, A is the actuator response, H is the sensor transfer function (constant factor),
X is the natural (unsuppressed) motion of MICH and S is an excitation injected at C1:LSC-MICH_EXC.
When the natural motion of MICH X is smaller than the excited displacement S/H, dividing MICH_OUT by S gives
[Transfer function] = S / MICH_OUT
= (1+G) /G * H * A
At low frequency the open loop gain is always big, so that the transfer function can be approximated to
[Transfer function] ~ H *A
This approximation is valid with a precision of 1/G.
In my case yesterday, the open loop gain at 0.1Hz was about 103 or more than that, so the uncertainty due to the loop gain was 0.1% or even less.
(Effect from the MICH motion)
In the equation [1], it is shown that the MICH motion X shows up together with the excitation signal.
Actually this MICH motion term was not completely negligible and eventually this term disturbs the measurement resulting in a low coherence.
In order to get a high coherence in the measurement, X should be smaller than the excited displacement S/H,
X << S / H
This the reason why I had to inject a big excitation signal. Although the coherence around 1Hz turned out to be still low due to the loud natural motion in MICH.
The excitation was already close to 0.1 um level in terms of peak-to-peak displacement, and I wasn't able to increase it any more because the MICH signal would run into a nonlinear regime.
In the worst case I lost the lock due to a too much excitation.
(Fringe hopping technique)
Actually I tried and gave up this technique. That's why I did the in-loop measurement.
My feeling is that this technique is not suitable for the 40m.
What I tried was to flip the sign of the MICH control such that the fringe hops from the dark fringe to the neighbor bright fringe or vice versa.
Difference in the control signal (C1:LSC-MICH_OUT) was supposed to give us the amount of signal which drives the actuator by exactly quarter of the laser wave length.
However this technique turned out to be not good because
(1) BS actuator is too strong
=> expected difference in the control signal is quite small.
=> \lambda / 4 / A ~ 12 counts, where A is the actuator DC response of about 2.2e-8 [m/counts].
(2) MICH motion was too loud
=> I saw such a tiny 12 counts difference in the control signal, but once the hopping is done the control signal immediately fluctuated and it was really hard to precisely measure it.
=> It's simply because MICH was loud, and the actuator tried to suppress the motion and it resulted such an immediate signal fluctuation in the control signal
|
This seems like an error prone method for DC responses due to the loop gain uncertainty. Better may be to use the fringe hopping method (c.f. Luca Matone) or the fringe counting method
|
|
5651
|
Tue Oct 11 17:32:05 2011 |
jamie | HowTo | Environment | 40m google maps link |
Here's another useful link:
http://maps.google.com/maps?q=34.13928,-118.123756 |
5650
|
Tue Oct 11 15:19:17 2011 |
rana | HowTo | Environment | 40m map |
The Kinemetrics dudes are going to visit us @ 1:45 tomorrow (Wednesday) to check out our stacks, seismos, etc.
 
I put these maps here on the elog since people are always getting lost trying to find the lab. |
5649
|
Tue Oct 11 15:14:50 2011 |
rana | Update | LSC | BS actuator reponse at low frequency : measured |
Quote: |
The response of the BS actuator in a low frequency regime has been measured.
|
This seems like an error prone method for DC responses due to the loop gain uncertainty. Better may be to use the fringe hopping method (c.f. Luca Matone) or the fringe counting method |
5648
|
Tue Oct 11 03:35:16 2011 |
kiwamu | Update | LSC | BS actuator reponse at low frequency : measured |
The response of the BS actuator in a low frequency regime has been measured.
After the measurement I did a coarse fit to see if the low frequency data agree with the high frequency response which I have measured two weeks ago ( #5583)
So far it shows a good agreement with the high frequency data (see the plot below). Tomorrow I will do a serious fitting.
Once the calibration of BS is done, the low frequency responses of ITMs, PRM and SRM will be done by simply exciting BS and comparing them (maybe at a couple of frequency points around 0.1Hz).
(Measurement)
+ With free swinging MICH, the sensor (AS55_Q) was calibrated into counts/m.
=> The peak-peak counts was about 110 counts. So the sensor response is about 6.5x108 counts/m
+ Locked Michelson with AS55_Q and the signal was fedback to BS.
+ Set the UGF high enough so that the open loop gain below 10 Hz is greater than 1.
+ With DDT's swept sine measurement, C1:LSC-MICH_EXC was excited with a big amplitude of 40 counts.
+ Took a transfer function from C1:LSC-MICH_OUT to C1:LSC-MICH_EXC.
+ Calibrated the transfer function into m/counts by dividing it with the sensor response.
Quote from #5641 |
One possible reason is that my actuator responses are not so accurate below 1Hz.
I will measure the DC response of all the actuators and it will completely determine the shapes of the actuator responses except for the region around the resonance.
|
|
5647
|
Tue Oct 11 00:59:55 2011 |
Suresh | Update | IOO | Preliminary locking of WFS loops |
[Kiwamu, Koji, Suresh]
After correcting several errors in the WFS loops, we turned them on today and saw them working!
A while back (last week actually) I noticed that the WFS1 and WFS2 QPD segments were numbered in a different order but that their input matrices did not reflect this change. As result the WFS pitch and yaw definitions were pretty much mixed up. However even after clearing this up the signals still showed significant amount of cross couplings.
This problem was finally traced to the relative phase between I and Q channels of the WFS segments. Koji suggested that I check the relative phase between all the segments to be sure. I then repeated the procedure that Valera and I followed in our earlier elog # 5321 , and found that the phases indeed required to be adjusted. The excitation of MCL was at 6Hz, 100mVpp, as before. The WFS response after this was much improved i.e. the pitch yaw cross couplings were not visible when we misalign the MC with sliders in MC_ALIGN. And the magnitude of the response also increased since the signal was transferred from the Q to I channels. The the phases were tweaked by hand till Q< 1% of I. However when I repeated this measurement an hour later (I wanted to save the plots) I found that the phases had changed by a few percent!
Koji noticed that the MC_REFL camera image showed significant intensity fluctuations and advised that we try a higher frequency and lower amplitude to avoid nonlinear effects in the WFS and in the MCL to PSL lock. So we repeated the process at 20Hz and 20mVpp, introduced at the IN2 of the MC_Servo. The fig below shows the level to which we reduced the signal in Q.

We then checked the relative phase between various quadrants by looking at the time series in dataviewer. WFS2 Seg4 phase had to be flipped to bring it into phase with all the rest.

After this I tried to see the WFS response to moving the MC1 and MC3 with the sliders and determined the following relations:
Pitch |
WFS1 |
WFS2 |
MC1 |
+ |
- |
MC2 |
- |
- |
MC3 |
+ |
+ |
Yaw |
WFS1 |
WFS2 |
MC1 |
+ |
+ |
MC2 |
- |
- |
MC3 |
+ |
- |
Disregarding the MC2 for now and assuming arbitrary gains of 1 for all elements we inverted these matrices inserted them into the WFS_servo_outmatrix. We then found that the with a sign flip on all elements the loops were stable. In the servo filters we had turned on only the filter modules 3 and 4. There was no low frequency boost. We gradually increased gain till we saw a significant suppression of the error signal at low frequencies as shown below. There was also an associated suppression of Intensity noise at REFL_DC after a single bounce from PRM.

To see if the locks can actually realign the MC if it were manually misaligned, we turned the loops off and misaligned MC by moving MC3 pitch by 0.05 (slider position), and then turned on the loops. The locks were reengaged successfully and the MC regained alignment as seen on the StripTool below:

We can now proceed with the fine tuning the servo filters and understand the system better:
Q1: Does the WFS (I to Q) phase drift rapidly? How can we prevent it?
Q2: How is that we do not see any bounce or roll resonances on the WFS error signals?
Q3: How do we include the MC2 QPD into the WFS Servo?
I will proceed with determination of the actual transfer coefs between the MC DoF and the WFS sensors.
|
5646
|
Mon Oct 10 18:53:04 2011 |
Katrin | Update | Green Locking | Mirrors whose angle of incidence is not 45° |
The angle of incidence of light is for some mirrors on the YARM end table different from 45° even though the mirrors are coated for 45°.
The mirrors below are useful if there are plans to replace these mirrors by properly coated ones.
Mirror
|
Angle of incidence (degree) |
1st 1" mirror right after laser* |
10 |
2nd 1" mirror right after laser |
35 |
1st 2" steering mirror to vacuum system |
15 |
2nd 2" steering mirror to vacuum system |
28 |
* This is the new mirror as decribed on http://nodus.ligo.caltech.edu:8080/40m/5623
|
5645
|
Mon Oct 10 16:32:18 2011 |
steve | Update | SUS | UL sensor of ETMY is recovered |
I lost UL osem voltage this morning when I was checking the actual connection at rack ETMY
This after noon I disconnected the 64 pins IDE connector from satelite amp at the rack, and the two 25 pins Dsubs at this juction board.
UL OSEM recovered after reconnecting these three connectors.
Atm3, bad connection.........noisy UL
|
5644
|
Mon Oct 10 15:41:56 2011 |
Koji | Update | PSL | PMC aligned |
[Koji Suresh]
The steering mirrors for PMC were aligned. The transmission went up from 0.779 to 0.852. |
5643
|
Mon Oct 10 13:52:04 2011 |
kiwamu | Update | LSC | RE: First attempt to estimate mode matching efficiency using interferometer |
Quote from #5640 |
"^2"s are missing in the second equation, but the calculation results seem correct.
PRX and PRY have different mode matching because of the Michelson asymmetry.
Are individually estimated mode matching indicates any sign of reasonable mode mismatch?
(The difference can be very small because the asymmetry is not so big.)
|
- Thank you for the correction. The missing square operation has been added correctly on the last entry (#5639).
- As for the individual MM efficiency,
I was assuming that the MM solutions are the same for PRX, PRY and the real PRC, so I haven't carefully checked differences between those cavities.
However as you mentioned the difference in those cavities can be tiny due to the small 3 cm Schnupp asymmetry.
Anyway I will briefly check it to make me sure. |
5642
|
Mon Oct 10 12:14:00 2011 |
Mirko | Update | Computer Scripts / Programs | IMC simulations |
[Mirko, Kiwamu]
I tried to answer two questions regarding the IMC:
1. What is the coupling of fluctuations in the SB freq. to SB transmitted power?
2. What (if any) is the influence of the IMC on the AM?
I ran into some weird things regarding the corresponding optickle simulations:
1. There seems to be some artifact at the beginning of every simulation sweep.
2. The position of features depends on the parameters of the sweep.
I mailed Matt asking if he sees some error in the simulations

|
5641
|
Mon Oct 10 10:14:43 2011 |
rana | Update | LSC | length fluctuations in SRCL |
How does it make sense that the motion at 0.1 Hz of PRC is 10x larger than MICH?
EDIT by KI:
That's actually the point which I was wondering at. One possible reason is that my actuator responses are not so accurate below 1Hz.
I will measure the DC response of all the actuators and it will completely determine the shapes of the actuator responses except for the region around the resonance.
In the process of producing the plot I was assuming that all the actuator response have a 1 Hz resonance with Q of 5.
However in reality this assumption is not true because the resonant frequency is different in each actuator. |
5640
|
Mon Oct 10 00:01:26 2011 |
Koji | Update | LSC | First attempt to estimate mode matching efficiency using interferometer |
"^2"s are missing in the second equation, but the calculation results seem correct.
PRX and PRY have different mode matching because of the Michelson asymmetry.
Are individually estimated mode matching indicates any sign of reasonable mode mismatch?
(The difference can be very small because the asymmetry is not so big.) |
5639
|
Sun Oct 9 17:13:46 2011 |
kiwamu | Update | LSC | First attempt to estimate mode matching efficiency using interferometer |
The efficiency of the mode matching (MM) to PRC (Power-Recycling Cavity) has been estimated by using the interferometer.
The estimated MM efficiency is about 74 % when losses in the cavity are assumed to be zero.
(Motivation)
There had been an issue that the recycling gain didn't go to the designed high value of about 42 ( #5541).
One of the possibilities is a low efficiency in the MM to PRC ( also see #5541).
Although the MM efficiency had been measured using a beam scan ( see a summary on the wiki) a long time ago, it haven't been verified.
Therefore the MM has to be reviewed by using the real interferometer.
(Measurement)
The concept of this measurement is observe the amount of the reflected light from a power-recycled cavity and estimate the MM efficiency based on the measured reflectivities.
Since using the real PRC (consisting of BS, ITMs and PRM) could be a too complicated system for this measurement,
simpler cavities, namely Power-Recycled ITMX and ITMY (PRX and PRY), were used to examine the MM efficiency.
The measurement goes in the following order :
(1) Measurement of the amount of the single-bounce reflection from PRM with BS and ITMs misaligned.
(2) Lock PRX or PRY to carrier resonance.
(3) Alignment of PRX/Y to maximize the intracavity power. This time ASDC was used as a monitor of the intracavity power.
(4) Measurement of the amount of the reflected light when the cavity is in resonance. The value in REFLDC was averaged in 100 sec.
=> done by tdsavg 100 C1:LSC-REFLDC_OUT
The same measurement was performed for both PRX and PRY.
- locking parameters -
Sensor = REFL11_I
Whitening gain = 10 (30 dB)
PRCL_GAIN = 2
UGF ~ 200 Hz
(Analysis)
In order to estimate the relation between the MM efficiency vs. the reflected light, two models are considered:
(1) simple model => no loss and no sidebands
(2) sideband-included model => no loss but sidebands are taken into the account of the reflection.
(1) In the simple model the reflectivity Prefl / Pin is expressed by
[Reflectivity] = Prefl / Pin = Z * Rcav + (1- Z) * Rprm
where Z is MM efficiency and Rprm is the reflectivity of PRM
and Rcav is the reflectivity of PRX/Y when it's resonance and it is defined by
R cav = | r prm - r itm t 2BS| 2 / |1 -r prm r itm t 2BS | 2
Tprm = 5.75% and Titm = 1.4 % are assumed in all the calculations.
In the first equation the first term represents the mode matched light and hence it couples with the cavity through Rcav.
The second term is the non-mode-matched light and because they are not interacting with the cavity they will be simply reflected by PRM through Rprm.
(2) In reality two phase-modulated light (11 MHz and 55 MHz) will behave differently from the carrier.
For example when the carrier is in resonance the sidebands will be anti-resonance against the cavity.
So that the amount of REFLDC will be slightly bigger because of the reflection of the sidebands.
Prefl = Z * Rcav * Pc + Z * Ranti * Ps + (1- Z) * Rprm * (Pc + Ps)
where Pc and Ps are the power in the carrier light and the sidebands respectively.
And Ranti is the reflectivity of the anti-resonance PRX/Y, which can be obtained by replacing the minus sign by the plus sign in the equation of Rcav shown above.
It is assumed that the sum of the carrier power and sidebands power is the incident power Pin = Pc + Ps.
The power in the carrier and the sidebands were estimated based on the OSA measurement ( #5519), so that
Pc / Pin = |J0(0.14)|2 * |J0(0.17)|2 = 0.976
Ps / Pin = 2 * |J1(0.14)|2 + 2 * |J1(0.17)|2 = 0.024
(Results)
Here are the measured values in REFLDC
-- Measurement 1 : PRX
Single bounce from PRM = 4802.27 counts
==> the incident power = 5095.25 counts
Reflected light from PRX = 4433.88 counts
==> Reflectivity = 0.8702
-- Measurement 2 : PRY
Single bounce from PRM = 4833.05 counts
==> the incident power = 5127.05 counts
Reflected light from PRX = 4444.48 counts
==> Reflectivity = 0.86672
On average the reflectivity of power-recycled ITM cavity was 0.868 with a standard deviation of 0.001744.
Actually the standard deviation estimated here is not fair because the measurement was done by only twice,
but my intention was that I wanted to see how the error can affect the estimation of the MM efficiency.
Here is a plot comparing the model curves and the measured values with 5 sigma error box (5 times of measured standard deviation).

It is shown that the mode matching efficiency is 73.7 % when the sideband-included model is considered.
With the 5 sigma deviation it can go from 65% to 82% but it is still low and lower than the beam scan measurement ( see a summary on the wiki).
Anyways the estimated MM efficiency with the sidebands effect included and without loss effect is
MM efficiency = 73.7 +/- 1.7 % (1 sigma error) or +/- 8.7 % (5 sigma error) |
5638
|
Sat Oct 8 04:41:07 2011 |
kiwamu | Update | LSC | length fluctuations in SRCL |
For a comparison, the length fluctuation of Signal-Recycled ITMX (SRX) and ITMY (SRY) have been measured.
Roughly speaking the length motion of SRX and SRY are as loud as that of PRCL.
Some details about the measurement and data analysis can be found in the past elog entry (#5582).
In the process of converting the raw spectra to the calibrated displacements the SRM actuator was assumed to have a resonance at 1Hz with Q = 5.

(Notes on SRX/Y locking)
Sensor = REFL11_I
Actuator = SRM
Demod. phase = 40 deg
SRCL_GAIN = 20
UGF = 100 - 200 Hz
Resonant condition = Carrier resonance
Whitening gain = 0 dB
ASDC = 360 counts
|
The MICH and PRCL motions have been measured in some different configurations.
+ PRCL is always noisier than MICH.
|
|
5637
|
Sat Oct 8 00:44:42 2011 |
kiwamu | Update | LSC | calibration of SRM actuator |
The AC response of the SRM actuator has been calibrated.
(Summary of the calibration results)
BS = 2.190e-08 / f2 [m/counts]
ITMX = 4.913e-09 / f2 [m/counts]
ITMY = 4.832e-09 / f2 [m/counts]
PRM = 2.022e-08 / f2 [m/counts]
SRM = 2.477e-08 / f2 [m/counts] ( NEW ! )
(Measurement)
The same technique as I reported some times ago ( #4721) were used.
The Signal-Recycled ITMY was locked for measuring the actuator response.
Since the ITMY actuator had been already calibrated, first the sensor was calibrated into [counts/m] by exciting the ITMY actuator and then calibrated the SRM actuator with swept sine measurement.
- - notes to myself
SRCL GAIN = 2.2
Sensor = REFL11_I
Demod. phase = 40 deg
Resonant condition = Carrier resonant
Gain in WF = 0 dB
Quote from #5583 |
The AC responses of the BS, ITMs and PRM actuators have been calibrated.
|
|
5636
|
Fri Oct 7 23:23:05 2011 |
kiwamu | Update | SUS | SRM oplev was oscillating |
The SRM oplevs were found to be oscillating because of a small phase margin.
I reduced the gains to the half of them. The peak which Steve found today maybe due to this oscillation.
Quote from #5630 |
The SRM bounce peak 18.33 Hz. Suresh helped me to retune filter through Foton, but we failed to remove it.
|
|
5635
|
Fri Oct 7 22:48:26 2011 |
Suresh | Update | IOO | MC2 Trans QPD spot size and incident power decreased |
After centering the spot on the MC2, I started to adjust the spot position on MC_TRANS_QPD to center the beam on it. I noticed that the spot size was about 3 to 4mm dia. because the 200mm lens was too close to the QPD. I moved it back and decreased the spot size to about 1mm and the sensitivity to spot position increased. However, Koji noted that the QPD sectors were near saturation, so I put in a ND=0.3 filter to reduce the incident power on the QPD.
At optimal alignment the current QPD_SUM is around 25k to 26k counts (factor of 2 down). Eventually the gain of the QPD ckts have to be reduced to prevent saturation, for the moment this is temporary fix.
The MC_TRANS_SUM trigger for MC autolocker is working fine no further change was required. |
5634
|
Fri Oct 7 22:41:05 2011 |
Koji | Configuration | General | Script backup fixed |
Script backup regularly runs on op340m by crontab,
but the true backup were not taken since Oct 16, 2010
as the backup program was looking at the old script directory.
/cvs/cds/script/backupScripts.pl was modified to look at the new script directory.
OLD COMMAND:
$command = "cd /cvs/cds/caltech; /usr/sbin/tar cfX - $EXCLUDE_LIST scripts | bzip2 > $ARCHIVEDIR/scripts_$curdate.tar.bz2";
NEW COMMAND:
$command = "cd /opt/rtcds/caltech/c1; /usr/sbin/tar cfX - $EXCLUDE_LIST scripts | bzip2 > $ARCHIVEDIR/scripts_$curdate.tar.bz2"; |
5633
|
Fri Oct 7 22:31:53 2011 |
kiwamu | Update | SUS | noisy ULSEN on ETMY |
Yesterday Koji was claiming that the rms monitor of ULSEN on ETMY didn't well go down.
Indeed something bad is happening on ULSEN of ETMY.
I guess it could be a loose connection.

(The unit of Y axis in the plot is not true. Don't believe it !) |
5632
|
Fri Oct 7 19:06:46 2011 |
Suresh | Update | IOO | MC spot positions: checked and corrected. |
Koji and Kiwamu had adjusted the MC beam axis slightly such that we can couple the MC output into the Y-arm without exceeding the current range of adjustability on PZT1. This changed the centering of beam spots on MC mirrors. I checked the mc-decentering make sure we have not made too big a compromise. And since we can move MC2 spot position while maintaining the current positions on MC1 and MC3 decentering, we can atleast eliminate the A2L coupling on that mirror. I used the scripts in $scripts$/MC/moveMC2/ to adjust the MC2 spot position.
Spot positions in mm (MC1,2,3 pit MC1,2,3 yaw) before adjustment:
1.4674 -0.3548 1.0199 -1.5519 1.9834 -1.5971
After correcting MC2:
1.4528 0.1431 0.9958 -1.2147 0.3823 -2.0163
After correcting MC1:
1.3745 0.0669 0.8899 -1.5269 0.0296 -1.7314
The spot positions on MC1 and MC3 are very nearly (+/- 0.06 mm) same as before, while the MC2 decentering has been reduced close to zero.
A slight adjustment of the PZTs may be required to reset the beam pointing. |
5631
|
Fri Oct 7 17:35:26 2011 |
Katrin | Update | Green Locking | Power on green YARM table |
After all realignment is finished, here are the powers at several positions:

|
5630
|
Fri Oct 7 14:04:48 2011 |
steve | Update | SUS | He/Ne intensity noise effect on oplevs |
The SRM bounce peak 18.33 Hz. Suresh helped me to retune filter through Foton, but we failed to remove it.
ITMX_OPLEV_PERROR shows high coherence with oplev laser. This is our lousiest set up. I will work on it next week.
Generally speaking we can say that JDSU-Uniphase 1103P and 1125P laser intensity noise do not limit oplev servo performance.
However, the overall well being of filters, gain settings, beam sizes on qpds are poor.
|
5629
|
Fri Oct 7 11:53:47 2011 |
Koji | Update | LSC | DRMI locked and some plans |
- REFL165 PD to be fixed (shows constant high voltage at the DC out)
- Make POP22/110 PD
- Install AS11? or use it as POX11?
- Install POP55 |
5628
|
Fri Oct 7 11:45:24 2011 |
Koji | Summary | LSC | POY11 installed, 55MHz PD at POY removed |
POY11 PD was installed last night. The lock of the Y arm was confirmed with the POY11I signal.
- The DC transimpedance was modified to be 1010V/A as the incident power is tiny.
- The demodulation phase of the roughly adjusted (148deg) to have PDH signal at the I-phase.
The comparison with AS55I signal exhibits that POY11I have ~150 times weaker signal with 45dB whitening.
(In total 25000 times weaker.)
On the way to make POY11 functioning, there were many fixes at the LSC rack...
Details:
- The PD interface cards (power supply for the RFPDs) were checked:
So far the two card at the right hand side were checked.
Desipite the previous entry reported the issues on those boards, they did not show any problem yesterday.
One hypothetical possibility is the enabling switches that is controlled from the old slow epics targets.
- POY55 was removed
This 55MHz PD is supposed to be installed at POP.
The PD, an RF cable, an RF amp, the power supply of the RF amp were removed.
- POY11 was installed
The PD was placed where the 55MHz was placed.
The beam was aligned on the diode using the IR viewer and the digital multimeter.
The power supply cable and the RF cable for POY on the ITMY table were used.
There were an ND filter on the POY beam path. It was removed.
- On the LSC rack
The PD RF was connected to the patch panel at the top of the rack.
There were loose connectors on the patch panel. Some connectors were tightened on the panel.
I found that POY11 and POX11 had I&Q signal reversely connected to the whitening board.
==> These were fixed but require the orthogonality test again for those channels.
The I phase output of the AS11 demod board had a broken connector.
The onboard SMA has got disintegrated because of too much twist on the connector.
The board was once removed from the rack and the connector was fixed using a heat gun and soldering.
The DC signals were checked. POYDC was not correctly connected. POYDC were correctly connected to the POYDC channel.
- CDS
c1lsc was found with the RFM frozen.
The c1lsc machine was soft-rebooted after stopping all of the RT processes.
Once the RT processes came back, they were all burtrestored.
- PDH locking
Restored Y-arm. Locked it with AS55Q.
Ran ASS alignment for Y-arm.
100cnt 150Hz sinusoidal signal is applied to ETMY
Measured the PSD of AS55Q, POY11I, and POY11Q.
Adjusted the demod phase so that the excitation could be minimized in POY11Q.
|
5627
|
Fri Oct 7 04:42:24 2011 |
kiwamu | Update | LSC | DRMI locked and some plans |
DRMI has been locked using the same RFPD selection as the old days (i.e. AS55_I, AS55_Q and REFL_I).(#4760)
But remember : this is just a beginning of several measurements and tests to characterize the central part.
Here is a list of the measurements and actions :
- 3f locking related
+ Listing up the necessary RFPDs and their installations.
+ Calibration of the SRM actuator => this is necessary to convert the sensing matrix into unit of [counts/m] or [W/m].
+ Measurement of the sensing matrix => check the performance of 3f signals. Also diagonalization of the LSC sensing matrix
+ Diagonalization of the output matrix.
+ Noise characterization of 3f PDs => confirm the noise are low enough to keep the lock of the central part
- Power-recycling gain issue related (#5541)
+ Estimation of the mode matching efficiency => maybe we can use power-recycled ITMs to estimate it (?)
+ Implementation of auto alignment servos and scripts for MICH, PRCL and SRCL. => integrate it to the existing ASS model
+ Search for a possible loss factor |
5626
|
Thu Oct 6 15:40:57 2011 |
Jenne | Update | LSC | Arm absl length data taken |
[Katrin, Jenne]
We took the data for the new absolute length measurement of both arms, after the latest vent and move. We will analyze soonly. We had done a round of analysis, but then Koji pointed out that our data wasn't so clean because the whitening filters were on (and saturated the ADC). We now have the data (but not the analysis) for the better data with the WF off.
So our dirty-data preliminary number for the X arm is 37.73meters, which is 14cm different from our old length. We were supposed to move by ~20cm, so....either this measurement is bad because the data sucked (which it did), or we are 6cm off. Or both.
I'll do another analysis with the clean data for both arms later today/tomorrow. |
5625
|
Thu Oct 6 15:37:26 2011 |
Jenne | Update | Green Locking | Y-green Mech Shutter Button |
[Katrin, Jenne]
We were poking around and tried to make a button for the Y-green shutter, just like the X-green already has. I don't know where the X-green shutter button goes to in model-land, so I can't figure out if there is already a channel set up for the Y end. Just switching the X for a Y didn't work. Someone (maybe me) should fix this in the next soon. |
5624
|
Thu Oct 6 05:18:20 2011 |
kiwamu | Update | LSC | Noise in AS55 was from clipping : fixed |
It turned out the noise in AS55 was due to a clipping. After fixing the clipping the noise successfully went down.
I was going to briefly check the clipping and go ahead locking DRMI, but for some reason I couldn't stop myself from working on this issue.
Here is a plot of the noise spectra taken before and after fixing the clipping.
The configuration of this measurement is exactly the same as that I did before (#5595)

(what I did)
+ Locked power-recycled ITMY so that the AS beam is bright enough to work with.
+ Shook BS at 1 Hz in the YAW direction
+ Looked around the AP table with an IR viewer and searched for a clipping moving at 1Hz.
+ Found the first lens in the AS beam path has clipped the beam at the upper side. A tiny portion of the beam was clipped.
+ Corrected the beam height to 4 inch by steering the very first mirror.
+ Raised the height of the lens because it was about 3.5 inch or so.
+ Found the lens had a scratch (~1 mm size ) at 1 cm blow the center on the surface.
=> I tried finding a spare 2 inch lens with a long focul length, but I couldn't find it,
So I left the lens as it is, but we should buy some 2 inch lenses just in case like this.
+ Replaced the 1 inch beam splitter by 2-inch 99% BS so that most of the light goes into the RFPD and a little bit goes into the camera.
Quote from #5595 |
The AS55 signal contains more noise than the REFL signals.
|
|
5623
|
Wed Oct 5 18:31:02 2011 |
Katrin | Update | Green Locking | Exchanged mirror on YARM table |
On the Green YARM end table the second mirror behind the laser has been exchanged.
Reason: The light is impinging on the mirror under an angle of about 10 degrees, but the old mirror was coated for angle of incidence (aoi) of 45°.
Thus it was more like a beam splitter. The new mirror is a 1" Goock & Housego mirror which has a better performance for an aoi of 10 degree.
Realignment through Faraday Isolator and SHG cristall as well as 532nm isolator is almost finished. |
5622
|
Wed Oct 5 17:08:49 2011 |
steve | Update | SUS | BS oplev spectra |
Kiwamu and Steve,
The He/Ne oplev shows no coherece so relative intensity noise is not limiting factor for the oplev servo |
5621
|
Wed Oct 5 14:18:09 2011 |
kiwamu | Update | CDS | some DAQ channel lost in c1sus : fb, c1sus and c1pem restarted |
I found again the ini files had been refreshed.
I ran the activateDQ.py script (link to the script wiki page) and restarted the daqd process on fb.
The activateDQ.py script should be included into the recompile or rebuild scripts so that we don't have to run the script everytime by hands.
I am going to add this topic on the CDS todo list (wiki page).
Quote from #5561 |
Somehow some DAQ channels for C1SUS have disappeared from the DAQ channel list.
|
|
5620
|
Wed Oct 5 11:33:25 2011 |
steve | Update | SUS | PRM and BS oplev laser replaced |
JDSU 1103P died after 4 years of service. It was replaced with new identical head of 2.9 mW output. The power supply was also changed.
The return spots of 0.04 mW 2.5 mm diameter on qpds are BS 3,700 counts and PRM 4,250 counts.
|
5619
|
Tue Oct 4 20:34:20 2011 |
Katrin | Update | Green Locking | 7kHz Peak in servo input YARM |
[Kiwamu, Katrin]
As reported earlier an oscillation around 7kHz is an the PDH error signal. The lower spectrum show that there is a peak from 6-7kHz.
This peak is somehow dependent on the modulation frequency. This means the peak can be shifted to a higher frequency when the modulation frequency is increased (see for comparsion f_mod=279kHz).
If the power supply for the green PD is switched of the peak vanishes. The same happens if the LO is switched of.

|
5618
|
Tue Oct 4 19:31:17 2011 |
kiwamu | Update | SUS | PRM and BS oplev laser died |
The He-Ne laser which has been used for the PRM and BS oplevs were found to be dead. 
According to the trend data shown below, it became dead during the dolphin issue.
(During the dolphin issue the output from the oplev QPDs are digitally zero)

|
5617
|
Tue Oct 4 19:06:46 2011 |
Koji | Update | IOO | closed the shutter before the MC |
Finished the work at 6:30
MC REFL (INLOCK): 0.50-0.52
MC REFL (UNLOCK): 6.9
MC TRANS: 54400~547000
RFAM level
Before the work: -48.5dBm for 1.07VDC (both 50Ohm terminated)
Right after the work: -80dBm for 0.896VDC (both 50Ohm terminated)
10min after: -70dBm
1hour after: -65dBm
3hours after: -62dBm
1day after (Oct 5, 20:00): -62.5dBm
2days after (Oct 6, 23:20): -72.5dBm
3 days after (Oct 7, 21:00): -57.8dBm
Quote:
|
The shutter before the MC was closed at 3:30 as I started working on the RFAM.
MC REFL (INLOCK): 0.6~0.7
MC REFL (UNLOCK): 6.9
MC TRANS: 50000~52000
|
|