  6628   Wed May 9 01:14:44 2012 JenneUpdateCDSNo signals for DTT from SUS

Quote:

UPDATE UPDATE:  Genius me just checked the FE status screen again.  It was fine ~an hour ago when I sat down to start interferometer-izing for the night, but now the SUS model and both of the ETMY computer models are having problems connecting to the fb.  *sigh* 

 Restarted SUS model - it's now happy. 

c1iscey is much less happy - neither the IOP nor the scy model are willing to talk to fb.  I might give up on them after another few minutes, and wait for some daytime support, since I wanted to do DRMI stuff tonight.

Yeah, giving up now on c1iscey (Jamie....ideas are welcome).  I can lock just fine, including the Yarm, I just can't save data or see data about ETMY specifically.  But I can see LSC data, so I can lock, and I can now take spectra of corner optics.

  6627   Wed May 9 00:45:13 2012 JenneUpdateCDSNo signals for DTT from SUS

Upgrades suck.  Or at least making everything work again after the upgrade.

On the to-do list tonight:  look at OSEM sensor and OpLev spectra for PRM, when PRMI is locked and unlocked.  Goal is to see if the PRM is really moving wildly ("crazy" as Kiwamu always described it) when it's nicely aligned and PRMI is locked, or if it's an artifact of lever arm between PRM and the cameras (REFL and AS).

However, I can't get signals on DTT.  So far I've checked a bunch of signals for SUS-PRM, and they all either (a) are just digital 0 or (b) are ADC noise.  Lame.

Steve's elog 5630 shows what reasonable OpLev spectra should look like:  exactly what you'd expect.

Attached below is a small sampling of different SUS-PRM signals.  I'm going to check some other optics, other models on c1sus, etc, to see if I can narrow down where the problem is.  LSC signals are fine (I checked AS55Q, for example).

UPDATE:  SRM channels are same ADC noise.  MC1 channels are totally fine.  And Den had been looking at channels on the RFM model earlier today, which were fine.

ETMY channels - C1:SUS-ETMY_LLCOIL_IN1 and C1:SUS-ETMY_SUSPOS_IN1 both returned "unable to obtain measurement data".  OSEM sensor channels and OpLev _PERROR channel were digital zeros.

ETMX channels were fine

UPDATE UPDATE:  Genius me just checked the FE status screen again.  It was fine ~an hour ago when I sat down to start interferometer-izing for the night, but now the SUS model and both of the ETMY computer models are having problems connecting to the fb.  *sigh* 

Attachment 1: OpLevs_OSEMs_allSUSchannels_ADCnoiseOnly.pdf
OpLevs_OSEMs_allSUSchannels_ADCnoiseOnly.pdf
Attachment 2: Screenshot.png
Screenshot.png
  6626   Tue May 8 17:48:50 2012 JenneUpdateCDSOAF model not seeing MCL correctly

Den noticed this, and will write more later, I just wanted to sum up what Alex said / did while he was here a few minutes ago....

 

Errors are probably really happening.... c1oaf computer status 4-bit thing:  GRGG.  The Red bit is indicating receiving errors.  Probably the oaf model is doing a sample-and-hold thing, sampling every time (~1 or 2 times per sec) it gets a successful receive, and then holding that value until it gets another successful receive. 

Den is adding EPICS channels to record the ERR out of the PCIE dolphin memory CDS_PART, so that we can see what the error is, not just that one happened.

Alex restarted oaf model:  sudo rmmod c1oaf.ko, sudo insmod c1oaf.ko .  Clicked "diag reset" on oaf cds screen several times, nothing changed.  Restarted c1oaf again, same rmmod, insmod commands.

Den, Alex and I went into the IFO room, and looked at the LSC computer, SUS computer, SUS I/O chassis, LSC I/O chassis and the dolphin switch that is on the back of the rack, behind the SUS IO chassis.  All were blinking happily, none showed symptoms of errors.

Alex restarted the IOP process:  sudo rmmod c1x04, sudo insmod c1x04.  Chans on dataviewer still bad, so this didn't help, i.e. it wasn't just a synchronization problem.  oaf status: RRGG. lsc status: RGGG. ass status: RGGG.

sudo insmod c1lsc.ko, sudo insmod c1ass.ko, sudo insmod c1oaf.ko .  oaf status: GRGG. lsc status: GGGG. ass status: GGGG.  This means probably lsc needs to send something to oaf, so that works now that lsc is restarted, although oaf still not receiving happily.

Alex left to go talk to Rolf again, because he's still confused.

Comment, while writing elog later:  c1rfm status is RRRG, c1sus status is RRGG, c1oaf status is GRGG, both c1scy and c1scx are RGRG.  All others are GGGG.

  6625   Tue May 8 16:43:15 2012 JenneUpdateCDSDegenerate channels, potentially a big mess

Rana theorized that we're having problems with the MC error signal in the OAF model (separate elog by Den to follow) because we've named a channel "C1:IOO-MC_F", and such a channel already used to exist.  So, Rana and I went out to do some brief cable tracing.

MC Servo Board has 3 outputs that are interesting:  "DAQ OUT" which is a 4-pin LEMO, "SERVO OUT" which is a 2-pin LEMO, and "OUT1", which is a BNC->2pin LEMO right now.

DAQ OUT should have the actal MC_F signal, which goes through to the laser's PZT.  This is the signal that we want to be using for the OAF model.

SERVO OUT should be a copy of this actual MC_F signal going to the laser's PZT.  This is also acceptable for use with the OAF model.

OUT1 is a monitor of the slow(er) MC_L signal, which used to be fed back to the MC2 suspension.  We want to keep this naming convention, in case we ever decide to go back and feed back to the suspensions for freq. stabilization.

Right now, OUT1 is going to the first channel of ADC0 on c1ioo.  SERVOout is going to the 7th channel on ADC0.  DAQout is going to the ~12th channel of ADC1 on c1ioo.  OUT1 and SERVOout both go to the 2-pin LEMO whitening board, which goes to some new aLIGO-style ADC breakout boards with ribbon cables, which then goes to ADC0.  DAQout goes to the 4pin LEMO ADC breakout, (J7 connector) which then directly goes to ADC1 on c1ioo.

So, to sum up, OUT1 should be "adc0_0" in the simulink model, SERVOout should be "adc0_6" on the simulink model, and DAQout should be "adc1_12" (or something....I always get mixed up with the channel counting on 4pin ADC breakout / AA boards). 

In the current simulink setup, OUT1 (adc0_0) is given the channel name C1:IOO-MC_F, and is fed to the OAF model.  We need to change it to C1:IOO-MC_L to be consistent with the old regime.

In the current simulink setup, SERVOout (adc0_6) is given the channel name C1:IOO-MC_SERVO.  It should be called C1:IOO-MC_F, and should go to the OAF model.

In the current simulink setup,DAQout (~adc1_12) doesn't go anywhere.  It's completely not in the system.  Since the cable in the back of this AA / ADC breakout board box goes directly to the c1ioo I/O chassis, I don't think we have a degenerate MC_F naming situation.  We've incorrectly labeled MC_L as MC_F, but we don't currently have 2 signals both called MC_F.

Okay, that doesn't explain precisely why we see funny business with the OAF model's version of MCL, but I think it goes in the direction of ruling out a degenerate MC_F name.

Problem:  If you look at the screen cap, both simulink models are running on the same computer (c1ioo), so when they both refer to ADC0, they're really referring to the same physical card.  Both of these models have adc0_6 defined, but they're defined as completely different things.  Since we can trace / see the cable going from the MC Servo Board to the whitening card, I think the MC_SERVO definition is correct.  Which means that this Green_PH_ADC is not really what it claims to be.  I'm not sure what this channel is used for, but I think we should be very cautious and look into this before doing any more green locking.  It would be dumb to fail because we're using the wrong signals.

 

Attachment 1: DegenerateIOOadc0chan6.png
DegenerateIOOadc0chan6.png
  6624   Tue May 8 10:43:42 2012 DenUpdateCDSbiquad filter form

Quote:

Quote:

I wanted to switch the implementation of IIR_FILTER from DIRECT FORM II to BIQUAD form in C1IOO and C1SUS models. I modified RCG file /opt/rtcds/rtscore/release/src/fe/controller.c by adding #define CORE_BIQUAD line:

#ifdef OVERSAMPLE
#define CORE_BIQUAD      
#if defined(CORE_BIQUAD)

 I am really not ok with anyone modifying controller.c.  If we're going to be messing around with that we need to change procedure significantly.  This is the code that runs all the models, and we don't currently have any way to track changes in the code.

Did you change it back?  If not, do so immediately and stop messing with it.  Please consult with us first before embarking on these kinds of severe changes to our code.  This is the kind of shit that other people have done that has bit us in the ass in the past.

Futhermore, there is already a way to enable biquad filters in the new version with out modifying the RCG source.  All you need to do is set biquad=1 in the cdsParameters block for you model.

DO NOT MESS WITH CONTROLLER.C!

 ok

  6623   Tue May 8 09:58:17 2012 DenUpdateCDSSUS -> FB

 [Alex, Den] 

It was in vain to restart mx_stream yesterday as C1SUS did not see FB

controls@c1sus ~ 0$ /opt/open-mx/bin/omx_info 

Open-MX version 1.3.901
 build: root@fb:/root/open-mx-1.3.901 Wed Feb 23 11:13:17 PST 2011
Found 1 boards (32 max) supporting 32 endpoints each:
 c1sus:0 (board #0 name eth1 addr 00:25:90:06:59:f3)
   managed by driver 'igb'
Peer table is ready, mapper is 00:60:dd:46:ea:ec
================================================
  0) 00:25:90:06:59:f3 c1sus:0
  1) 00:60:dd:46:ea:ec fb:0                           // this line was missing
  2) 00:14:4f:40:64:25 c1ioo:0
  3) 00:30:48:be:11:5d c1iscex:0
  4) 00:30:48:bf:69:4f c1lsc:0
  5) 00:30:48:d6:11:17 c1iscey:0
 
At the same time FB saw C1SUS:
 
controls@fb ~ 0$ /opt/mx/bin/mx_info
 
MX Version: 1.2.12
MX Build: root@fb:/root/mx-1.2.12 Mon Nov  1 13:34:38 PDT 2010
1 Myrinet board installed.
The MX driver is configured to support a maximum of:
8 endpoints per NIC, 1024 NICs on the network, 32 NICs per host
===================================================================
Instance #0:  299.8 MHz LANai, PCI-E x8, 2 MB SRAM, on NUMA node 0
Status: Running, P0: Link Up
Network: Ethernet 10G

MAC Address: 00:60:dd:46:ea:ec
Product code: 10G-PCIE-8AL-S
Part number: 09-03916
Serial number: 352143
Mapper: 00:60:dd:46:ea:ec, version = 0x00000000, configured
Mapped hosts: 6

                                                        ROUTE COUNT
INDEX    MAC ADDRESS     HOST NAME                        P0
-----    -----------     ---------                        ---
   0) 00:60:dd:46:ea:ec fb:0                              1,0
   1) 00:30:48:d6:11:17 c1iscey:0                         1,0
   2) 00:30:48:be:11:5d c1iscex:0                         1,0
   3) 00:30:48:bf:69:4f c1lsc:0                           1,0
   4) 00:25:90:06:59:f3 c1sus:0                           1,0
   5) 00:14:4f:40:64:25 c1ioo:0                           1,0
 
For that reason when I restarted mx_stream on c1sus, the script tried to connect to the standard 00:00:00:00:00:00 address, as the true address was not specified.
 
Alex restarted mx on FB. Note, DAQD process will not allow one to do that until it runs, at the same time, you can't just kill it, it will restart automatically. For that reason one should open /etc/inittab and replace respawn to stop in the line
 
daq:345:respawn:/opt/rtcds/caltech/c1/target/fb/start_daqd.inittab
 
then execute inittab using init q and restart mx on the FB
 
controls@fb ~ 0$ sudo /sbin/init q
controls@fb ~ 0$ sudo /etc/init.d/mx restart
 

After that C1SUS started to communicate with FB. But the reason why this happened and how to prevent from this in future Alex does not know.

Restarting DAQD process (or may be C1SUS) also solved the problem with guralp channels, now they are fine. Again, why this happened is unknown.

 

  6622   Tue May 8 09:47:53 2012 JamieUpdateCDSbiquad filter form

Quote:

I wanted to switch the implementation of IIR_FILTER from DIRECT FORM II to BIQUAD form in C1IOO and C1SUS models. I modified RCG file /opt/rtcds/rtscore/release/src/fe/controller.c by adding #define CORE_BIQUAD line:

#ifdef OVERSAMPLE
#define CORE_BIQUAD      
#if defined(CORE_BIQUAD)

 I am really not ok with anyone modifying controller.c.  If we're going to be messing around with that we need to change procedure significantly.  This is the code that runs all the models, and we don't currently have any way to track changes in the code.

Did you change it back?  If not, do so immediately and stop messing with it.  Please consult with us first before embarking on these kinds of severe changes to our code.  This is the kind of shit that other people have done that has bit us in the ass in the past.

Futhermore, there is already a way to enable biquad filters in the new version with out modifying the RCG source.  All you need to do is set biquad=1 in the cdsParameters block for you model.

DO NOT MESS WITH CONTROLLER.C!

  6621   Tue May 8 03:38:28 2012 JenneUpdateGeneralMEDM svn emergency!!!

We officially are *failures* at svn-ing our scripts and screens.  This is NOT OKAY.  I checked in a few things, since there were already folders on the svn, but many things don't have folders created.  It's a hot mess.  We need to get our shit together, and become as disciplined about MEDM and scripts as we have been (under Jamie's watchful eye) of the simulink models.

I'm not going to start fixing it all right now.  It might not even happen at this point until after GWADW, but it needs to happen.

  6620   Tue May 8 03:33:24 2012 JenneUpdateSUSNew misalign / restore scripts for IFO_ALIGN screen

Since Den wasn't able to fix c1sus (make it talk to the framebuilder) before he left a few hours ago, I decided to do some housekeeping rather than actual locking. 

I wrote new save / misalign / restore scripts for all of the suspended optics on the C1IFO_ALIGN screen.  Adding save / restore versions for the mode cleaner optics should be quick and easy.  Now when you use the ! button for each optic, it points you to the new scripts.  I still have the burt capabilities there, but the restore script has the burt-restore line commented out.

Functionality:

SAVE:  burt-save the PIT_COMM and YAW_COMM values, as well as write those values and the date to a text file.

MISALIGN:  Turn off oplevs, move 100 steps of 0.01 in the "+" direction.

RESTORE: Move ~100 steps toward the saved value, until you're within 0.001 of the saved value (step size is "saved val" minus "current val" divided by 100).  Then just write the saved value to the slider (otherwise if the slider were touched between the last "save" and the restore, we might not be able to step precisely to the value we want). Turn oplevs back on.

Scripts are in the same place the old ones used to live:  ...../caltech/c1/medm/c1ifo/cmd/   New scripts are C1IFO_OPTIC(save/restore/misalign)_soft.cmd

I'm checking this one off of the to-do list.

 

Good things:  (a) I remembered / re-learned / just plain learned a lot about scripting.  (b) the optics are now walked slowly over to their misaligned state, and slowly walked back.  The past regime had the optics suddenly kicked over by a lot, sometimes enough to trip / come close to tripping watchdogs, which was never good.

Bad things: it took a long time.  Now it's bedtime.

  6619   Mon May 7 22:39:37 2012 DenUpdateCDSc1sus

[Jenne, Den]

We decided to reboot C1SUS machine in hope that this will fix the problem with seismic channels. After reboot the machine could not connect to framebuilder. We restarted mx_stream but this did not relp. Then we manually executed

/opt/rtcds/caltech/c1/target/fb/mx_stream -s c1x02 c1sus c1mcs c1rfm c1pem -d fb:0 -l /opt/rtcds/caltech/c1/target/fb/mx_stream_logs/c1sus.log

but c1sus still could not connect to fb. This script returned the following error:

controls@c1sus ~ 128$ cat /opt/rtcds/caltech/c1/target/fb/mx_stream_logs/c1sus.log


c1x02
c1sus
c1mcs
c1rfm
c1pem
mmapped address is 0x7fb5ef8cc000
mapped at 0x7fb5ef8cc000
mmapped address is 0x7fb5eb8cc000
mapped at 0x7fb5eb8cc000
mmapped address is 0x7fb5e78cc000
mapped at 0x7fb5e78cc000
mmapped address is 0x7fb5e38cc000
mapped at 0x7fb5e38cc000
mmapped address is 0x7fb5df8cc000
mapped at 0x7fb5df8cc000
send len = 263596
OMX: Failed to find peer index of board 00:00:00:00:00:00 (Peer Not Found in the Table)
mx_connect failed

Looks like CDS error. We are leaving the WATCHDOGS OFF for the night.

  6618   Mon May 7 21:46:10 2012 DenUpdateCDSguralp signal error

GUR1_XYZ_IN1 and GUR2_XYZ_IN1 are the same and equal to GUR2_XYZ.  This is bad since GUR1_XYZ_IN1 should be equal to GUR1_XYZ.  Note that GUR#_XYZ are copies of GUR#_XYZ_OUT, so there may be (although there isn't right now) filtering between the _IN1's and the _OUT's.  But certainly GUR1 should look like GUR1, not GUR2!!!

Looks like CDS problem, maybe some channel-hopping going on? I'm trying a restart of the c1sus computer right now, to see if that helps.....

Figure:  Green and red should be the same, yellow and blue should be the same.  Note however that green matches yellow and blue, not red.  Bad.

guralps.png

 

 

  6617   Mon May 7 21:19:00 2012 JenneUpdateSUSETMX oplev

I'm still not super happy with the low power level of the ETMX oplev, so I went to investigate.

OpLevETMX_1100days.png

This is a 3-year plot.  The first ~year in the plot, the oplev sum is ~15000 cts, and in the most recent year, it's ~1000 cts.  A new HeNe was installed in May 2011 (elog 4686), with an output of 2.6mW, after the old one had died.  When the new one was installed, Steve said that it was giving ~1400 counts, so maybe 1000 isn't too, too embarrassing.  There is, however, a lens on the injection side, which is AR coated for 1064.  The power before this lens (measured with the Ophir, set to 632nm) was ~2.6mW.  The power after this lens was ~1.5mW.  Now THAT is embarrassing.  I'm adding replacing that lens to the to-do list (elog 6595), although I don't want to do it until such a time (maybe in an hour, maybe in a few days?) when I've got the Xarm locked / aligned, so I can nicely re-center the oplev.  UPDATE:  The lens is a KBC 037 (-100) lens, and the sticker on the lens mount says coated for 1064.  We don't have any KBC037's in the visible lens kits, so we need to get one before I can do this replacement (PURCHASED 10pm).

There is an elog (elog 5004) from July 2011, which mentions that these channels have not been saved for a long time, so that's why there's the year-long gap. 

 

  6616   Mon May 7 21:05:38 2012 DenUpdateCDSbiquad filter form

I wanted to switch the implementation of IIR_FILTER from DIRECT FORM II to BIQUAD form in C1IOO and C1SUS models. I modified RCG file /opt/rtcds/rtscore/release/src/fe/controller.c by adding #define CORE_BIQUAD line:

#ifdef OVERSAMPLE
#define CORE_BIQUAD      
#if defined(CORE_BIQUAD)

C1IOO model compiled, installed and is running now. C1SUS model compiled, but during installation I've got an error:

controls@c1sus ~ 0$ rtcds install c1sus


Installing system=c1sus site=caltech ifo=C1,c1
Installing /opt/rtcds/caltech/c1/chans/C1SUS.txt
Installing /opt/rtcds/caltech/c1/target/c1sus/c1susepics
Installing /opt/rtcds/caltech/c1/target/c1sus
Installing start and stop scripts
/opt/rtcds/caltech/c1/scripts/killc1sus
Performing install-daq
Updating testpoint.par config file
/opt/rtcds/caltech/c1/target/gds/param/testpoint.par
/opt/rtcds/rtscore/branches/branch-2.5/src/epics/util/updateTestpointPar.pl -par_file=/opt/rtcds/caltech/c1/target/gds/param/archive/testpoint_120507_205359.par -gds_node=21 -site_letter=C -system=c1sus -host=c1sus
Installing GDS node 21 configuration file
/opt/rtcds/caltech/c1/target/gds/param/tpchn_c1sus.par
Installing auto-generated DAQ configuration file
/opt/rtcds/caltech/c1/chans/daq/C1SUS.ini
Installing EDCU ini file
/opt/rtcds/caltech/c1/chans/daq/C1EDCU_SUS.ini
Installing Epics MEDM screens
Running post-build script

ERROR: Could not find file: test.py
Searched path: :/opt/rtcds/userapps/release/cds/c1/scripts:/opt/rtcds/userapps/release/cds/common/scripts:/opt/rtcds/userapps/release/isc/c1/scripts:/opt/rtcds/userapps/release/isc/common/scripts:/opt/rtcds/userapps/release/sus/c1/scripts:/opt/rtcds/userapps/release/sus/common/scripts:/opt/rtcds/userapps/release/psl/c1/scripts:/opt/rtcds/userapps/release/psl/common/scripts
Exiting
make: *** [install-c1sus] Error 1

Jamie, what is this test.py?

  6615   Mon May 7 20:15:37 2012 DenUpdateIOOWFS noise in MC

Quote:

OK. Then we should make this number bigger such that the coherence is still completely maintained.
Is this set in the auto locker? Or manually set?

Quote:

  This effect is due to C1:IOO-WFS1_PIT_LIMIT=2000. When I turned if off, coherence between C1:IOO-WFS_PIT input and output signals restored

 

 

LIMIT is set manually, auto locker does not change it. I've put C1:IOO-WFS1_PIT_LIMIT=4000, it seems to be fine for now.

  6614   Mon May 7 19:39:52 2012 KojiUpdateIOOWFS noise in MC

OK. Then we should make this number bigger such that the coherence is still completely maintained.
Is this set in the auto locker? Or manually set?

Quote:

  This effect is due to C1:IOO-WFS1_PIT_LIMIT=2000. When I turned if off, coherence between C1:IOO-WFS_PIT input and output signals restored

 

  6613   Mon May 7 17:28:29 2012 DenUpdateIOOWFS noise in MC

Quote:

One filter bank adds noise to WFS signals and because of that we loose coherence between MC_F and seismic motion.

 This effect is due to C1:IOO-WFS1_PIT_LIMIT=2000. When I turned if off, coherence between C1:IOO-WFS_PIT input and output signals restored

wfs_limit.png

The other thing is that WFS actuate on the angular motion, but this couples to the position motion. We need to diagonalize the actuator.

  6612   Mon May 7 12:57:34 2012 DenUpdateIOOWFS noise in MC

I've measured coherence between seismometer signals and OSEMS of MC 1,2,3

gur_suspos_coh.png

GUR1 was rotated on the angle pi / 4 relative to x arm to match suspos axis of MC1 and MC3. Coherence between MC2 and GUR2_X is high, between MC3 and GUR1_X is ~0.2 but is compensated by GUR1_Y, between MC1 and GUR1_Y below 1Hz is ~0.5 and is not compensated by GUR1_X.

Then I measured coherence between GUR1_Y and MC3 OSEMS in 3 regimes :

  • FEEDBACK OFF, WFS OFF
  • FEEDBACK ON, WFS OFF
  • FEEDBACK ON, WFS ON

mc1_gur.png

Once WFS are ON, signal becomes noisy, FEEDBACK is OK. Then I measured coherence between MC_F and GUR2_X with WFS ON and OFF

mcl_wfs.png

Coherence between MC_F and GUR2_X is less when WFS are ON in the range 0.2 - 1 Hz and 4 - 10 Hz. Moreover PSD of MC_F seems to be higher at 10 - 100 Hz. But this may be caused by other reasons.

Then I measured the coherence between IN and OUT signals in WFS_SERVO

wfs_servo.png

One filter bank adds noise to WFS signals and because of that we loose coherence between MC_F and seismic motion.

  6611   Mon May 7 01:07:58 2012 DenUpdateIOOseismic in mcf

I tried to figure out where additional (to seismic) noise enters to MC_F, so the coherence below 1 Hz is low (~0.2-0.5). I've examined noise in the path

PD -> DEMOD -> MC BOARD -> FSS -> PZT Controller -> LASER

  • I terminated ADC and measured its noise
  • connected MC BOARD OUT to ADC, terminated INPUT1 of the MC BOARD and measured noise
  • connected DEMOD Q OUT to MC BOARD INPUT1, terminated PD INPUT on the DEMODULATOR and measured noise
  • connected PD to DEMOD, blocked the beam incident to the PD and measured noise

pd-mcf.jpeg

 MC BOARD noise shows up only below 0.1 Hz and at 10 Hz where the whitening filter starts to work. SNR is ~2 at 4 Hz, so we might want to slightly improve whitening filter. But other then that path PD->DEMOD->MC BOARD is not responsible for additional noises below 1 Hz.

Next I connected SR785 to the laser and measured the closed loop feedback signal while MC was locked.

fb.png     coh_err_fb.png

Coherence between MC BOARD OUT1 and feedback signal is high enough to assume that FSS and PZT controller are also not responsible for additional noise.

From the other hand MC2 SUSPOS measured by OSEMS shows good coherence with GUR1_X

mc2_suspos.jpg

That means that MC2 is indeed driven by seismic motion. In order to figure out if this is the case for MC1 and MC3, I rotated GUR2 by 45 degrees. When it will calm down, I'll measure coherence between OSEMs and seismic motion.

  6610   Sun May 6 01:41:55 2012 DenUpdatePEMseismic in x,y arms

I locked x,y arms and measured coherence between POS{X,Y}11_I_ERR, MC_F and seismometer signals.

coh_xyf.png

Surprisingly, coherence between POSY and GUR1Y is low, but with GUR1X is relatively high. I wonder if this is due to MC that brings this x-axis noise to the arms.

  6609   Sun May 6 00:11:00 2012 DenUpdateCDSmx_stream

c1sus and c1iscex computers could not connect to framebuilder, I restarted it, did not help. Then I restarted mx_stream daemon on each of the computers and this fixed the problem.

sudo /etc/init.d/mx_stream restart

  6608   Sat May 5 20:42:59 2012 DenUpdatePSLPMC

[Koji, Den]

Koji was right that I misaligned everything during the alignment work. I assumed that PMC should autolock and when I saw that it did not, I thought the laser is misaligned.

What we did:

1. Aligned mirrors to get the beam on the PD PMC REFL and PMCR camera. The PSL-PMC_RFPDDC was ~800 mV.

2. We disabled PMC servo, switching it to test position and changed "DC output adjust" by 0.01 in a loop

while true
do
    ezcawrite "C1:PSL-PMC_RAMP" -4.50
    ezcastep "C1:PSL-PMC_RAMP" "+0.01,450" -s "0.1"
    ezcawrite "C1:PSL-PMC_RAMP" 0.0
    ezcastep -s "0.1" -- "C1:PSL-PMC_RAMP" "-0.01,450"
done

3. While the script was running we adjusted the position of the beam on the far PMC mirror looking at an IR viewer. The goal is to align two steering mirrors to catch some resonances. We monitored them on the oscilloscope and on the PMCT camera.

4. We locked PMC and aligned steering mirrors.

  6607   Sat May 5 12:23:38 2012 KojiUpdatePSLPMC

No matter how you connect/disconnect, touching the laser may cause the PMC unlocked.

At least, I don't see the PMC reflection on the PD.
This means that the beam towards the PMC is largely misaligned.

If you are not sure what is misaligned, stop touching the table.
Close the shutter of the laser on the laser housing and leave the optics as they are.

  6606   Sat May 5 10:20:21 2012 DenUpdatePSLPMC

Quote:

I suspect that it was just unlocked when you had disconnected the cable.

There is not reflection now. It seems that it is now misaligned after the alignment work.

So what you need is "align while scanning PZT -> lock -> align".

Quote:

Unfortunately, after the measurement when I unplugged BNS connector from the laser, I misaligned PMC. For several hours I adjusted the mirrors but could not significantly improve transmitted signal. I'll return to this issue tomorrow.

 

 No, no, it was unlocked after I connected the cable back. The beam was even not on the PMC. I'll try PZT -> lock -> align.

  6605   Sat May 5 09:13:02 2012 KojiUpdatePSLPMC

I suspect that it was just unlocked when you had disconnected the cable.

There is not reflection now. It seems that it is now misaligned after the alignment work.

So what you need is "align while scanning PZT -> lock -> align".

Quote:

Unfortunately, after the measurement when I unplugged BNS connector from the laser, I misaligned PMC. For several hours I adjusted the mirrors but could not significantly improve transmitted signal. I'll return to this issue tomorrow.

 

  6604   Sat May 5 01:24:07 2012 DenUpdatePSLPMC

I was interested what whitening filter do we have between MC servo and ADC. The shape is in the figure below, SR provided 1 V white noise. Before the whitening filter MC_F is measured in Volts with SR and ADC (for ADC the shape is calculated using the whitening filter form):

whitening.jpg     mcf_v.png

I also wondered if FSS or PZT servo can add noise to the mode cleaner length signal and what is their gain. It should be big, as the laser's calibration is ~1 MHz/V => to account for seismic noise of 10^-6 m at 1 Hz, the voltage given to the laser should be ~ 1 V. And it is indeed the case. The gain is ~1000. I measured the coherence between MC_F and the laser fast input. It is 1 in the range measured (0.05 - 100 Hz). FSS and PZT do not add significant noise.

Unfortunately, after the measurement when I unplugged BNS connector from the laser, I misaligned PMC. For several hours I adjusted the mirrors but could not significantly improve transmitted signal. I'll return to this issue tomorrow.

  6603   Fri May 4 17:46:46 2012 JenneUpdateGreen LockingPSL doubling oven back on

While walking past the PSL, I noticed that the PSL's doubling oven's heater was still disabled from the power outage.  As with the ETMY heater, I hit the Enable button, and it started warming up (according to the front panel at least).

  6602   Fri May 4 17:44:42 2012 JenneUpdateGeneralOpLev 90-day trend

Quote:

After fixing the PRM tonight, all of the oplevs look okay. 

.....except ITMX, which looks like it's power dropped significantly after the CDS upgrade.  To be investigated tomorrow.

 I had a look-see at ITMX's oplev.  I can't see any clipping, so maybe the power is just low?  One thing that was funny though is the beam coming directly from the laser.  There is the main, regular beam, and then there is a thin horizontal line of red light also coming straight out of the laser.  I don't know what to do about that, except perhaps put an iris right after the HeNe, to block the horizontal part?  I'm not sure that it's doing anything bad to the optic though, since the horizontal part gets clipped by other optics before the beam enters the vacuum, so there is no real irregularity on the beam incident on the QPD.

I realigned the oplev on the QPD, using last night's ITMX alignment + whatever drift it picked up over night, so it may need re-recentering after Xarm is nicely aligned.

  6601   Thu May 3 22:37:44 2012 JenneUpdateGeneralOpLev 90-day trend

After fixing the PRM tonight, all of the oplevs look okay. 

.....except ITMX, which looks like it's power dropped significantly after the CDS upgrade.  To be investigated tomorrow.

Attachment 1: OpLevTrends_90days_Ending3May2012
  6600   Thu May 3 21:13:48 2012 KojiSummarySUSITMX/PRM/BS OPLEV aligned

[Jenne/Den/Koji]

We locked Xarm/Yarm and manually alignmed ITMX/ITMY/BS/ETMX/ETMY/PZT1/PZT2.

ITMY OPLEV was largely misaligned ==> The beam was centered on the QPD.

----

Then we aligned PRM using SB locking PRMI.

We noticed that OPLEV servo does not work. It made the PRM just noiser.

We went into the PRM table and found that the OPLEV beam was clipped in the vacuum chamber.
We tried to maximize the reflected beam from the window by touching the steering mirrors at the injection side.

Then the reflected beam was introduced to the center of the QPD.

After the alignment, the OPLEV QPD SUM increased to 4000ish from 200ish.
According to the OPLEV trend data, this is a nominal value of the QPD SUM.

Now the OPLEV servo does not go crazy.

 --

BS OPLEV beam was centered on the QPD.

  6599   Thu May 3 19:52:43 2012 JenneUpdateCDSOutput errors from dither alignment (Xarm) script

epicsThreadOnceOsd epicsMutexLock failed.
Segmentation fault
Number found where operator expected at -e line 1, near "0 0"
        (Missing operator before  0?)
syntax error at -e line 1, near "0 0"
Execution of -e aborted due to compilation errors.
Number found where operator expected at (eval 1) line 1, near "*  * 50"
        (Missing operator before  50?)
epicsThreadOnceOsd epicsMutexLock failed.
Segmentation fault
Number found where operator expected at -e line 1, near "0 0"
        (Missing operator before  0?)
syntax error at -e line 1, near "0 0"
Execution of -e aborted due to compilation errors.
syntax error at -e line 1, at EOF
Execution of -e aborted due to compilation errors.
Number found where operator expected at (eval 1) line 1, near "*  * 50"
        (Missing operator before  50?)
epicsThreadOnceOsd epicsMutexLock failed.
status : 0
I am going to execute the following commands
ezcastep -s 0.6 C1:SUS-ETMX_PIT_COMM +-0,50
ezcastep -s 0.6 C1:SUS-ITMX_PIT_COMM +,50
ezcastep -s 0.6 C1:SUS-ETMX_YAW_COMM +,50
ezcastep -s 0.6 C1:SUS-ITMX_YAW_COMM +-0,50
ezcastep -s 0.6 C1:SUS-BS_PIT_COMM +0,50
ezcastep -s 0.6 C1:SUS-BS_YAW_COMM +0,50
hit a key to execute the commands above

  6598   Thu May 3 17:15:38 2012 KojiUpdateLSCc1iscaux2 rebooted/burtrestored

[Jenne/Den/Koji]

We saw some white boxes on the LSC screens.
We found c1iscaux2 is not running.

Once the target was power-cycled, these epics channels are back.
Then c1iscaux2 were burtrestored using the snapshot at 5:07 on 4/16, a day before the power glitch.

  6597   Thu May 3 13:35:56 2012 JenneBureaucracyGeneral40m Meeting Action Items

Action Items:

IFO:

         Arm cavity sweeps, mode scan - JENNE

        Align AS OSA (others?) - JENNE       

        Investigate PRMI glitches, instability (take PRM oplev spectra locked, unlocked, to see if PRM is really moving) - KOJI, JENNE, DEN

         Connect up beatbox for Xarm use - KOJI with JENNE       

 

TT:

Prepare electronics for TTs (coil drivers) - JAMIE

In-air TT testing to confirm we can control / move TTs before we vent (starting in ~2 weeks) - SURESH

Connect TTs to digital system and controls, lay cables if needed - JAMIE with SURESH

OAF:

OAF comparison plot, both online and offline, comparing static, adaptive and static+adaptive - DEN

Static-only OAF noise budget (Adaptive noise budget as next step) - DEN

Build amplifiers for new small microphones - DEN 

SLC:

Black glass: to clean&bake - KOJI

Scattered light measurement at the end stations: design / confirmation of the mechanical parts/optics/cameras - JAN

MM:

IPPOS beam measurement - SURESH with JENNE

AS beam measurement (if beam is bright enough) - SURESH and JENNE

Mode matching calculations, sensitivity to MC waist measurement errors, PRM position - JENNE

Think up diagnostic measurement to determine mode matching to PRC while chambers are open, while we tweak MMT - JAMIE, JENNE, KOJI, SURESH

Computers:

       Upgrade Rossa, Allegra to Ubuntu, make sure Dataviewer and DTT work - JAMIE

 

  6596   Thu May 3 13:19:17 2012 JenneUpdateLockingMore success last night

I locked each arm, and the Michelson last night, no problems after I increased the Yarm gain from 0.1 to 0.2 .  I checked the green beam alignment just before going home, and both of the green beams are locking on ~03 or 04 modes, so aligning them is on the list for today.

  6595   Thu May 3 13:07:05 2012 KojiBureaucracyGeneralInterferometer check-up session

From 2PM, Jenne/Den/Koji. Finished at 10:30PM

  • General
    • More LCD screens! Wider screens!
      • Allegra - only can handle single head right now, but should prepare for the eventual upgrade of that videocard/whole computer and get 2 monitors.
      • Every workstation should have 2 24" monitors.  We currently have 2 good ones (one on Rossa, one on Pianosa), so we need 6.
    • Remove Control room shelf currently holding OSA 'scopes
    • Adjust both TV shelves so that they're at the same height, ~6" higher than the current left shelf.
      This gives space for stacking the 24" monitors for each computer.
    • Audio system for the signals!!!! Even a crappy one!
           /cvs/cds/ligo/centos5.5/cds/project/gds-2.14.2/Monitors/rockIFO
           /cvs/cds/gds/gds/Monitors/rockIFO
           (really /cvs/cds, not /opt)

  • IOO
    • PZT or Picomotor mounts for PSL/ALS beams
    • ALS on the both arm simultaneously / common/diff ALS scripts
  • SUS
    • PRM OPLEV servos are sometimes not stable. (fixed, JCD/DM/KA, 5/2/2012)
    • ETMX oplev - replace 1064nm lens with 632nm lens (KBC037, -100)
  • IFO
    • It seems that there is an occasional common-mode power transient in the arm transmissions.
      -> Track down the cause and fix.
    • Fix arm ASS even on CentOS
    • Drift of the green incident axis -> Assess the amount of the drift / replace the mount
    • Calibration of POP22 / AS110
    • PMC/IMC/ARM characterization (loss, finesse, reflectivity, etc)
  • CDS
    • Capture OSA signals in CDS (the 'scope TDS1001B has a USB port in the back for connecting to the computer)
    • Transmon (arms) for high and low power
    • POX11 whitening is not toggling the analog whitening???
    • CentOS machines cannot open simulink models properly (get "Bad Link"s everywhere, so you can't do anything useful).
    • Dataviewer and DTT can't connect to framebuilder from CentIOS machines
  • Electronics
    • Actuator noise level characterization (coil driver response in V/m & coil driver noise level V/rtHz)
    • Beat box
    • I/Q DFD
    • Improvement of POP22/110/AS110 RF circuits.
  • MEDM
    • OL alert in the watch dog screen (this is awesome!) should have small text saying "OL" (done, JCD, 5May2012)
    • Complete 40m overview screen - everything should be clickable with pseudo 3D icons
    • BETTER SUSPENSION SCREEN!!!
    • Script to generate a MEDM tree
    • Resurrect MEDM snapshots
       
    • The LSC screen has some "white" boxes -> investigate and fix them (done, KA, 5/2/2012)
    • Make the LSC control screen (compact) in the vertical arrangement (done, KA, 5/2/2012)
    • The signals of the watch dog screen should go red if any of the WDs are not enabled.
      =>Copy the logic of the signals in LSC screen.
      (done, KA, 5/2/2012)
    • Remove c1gfd from the cds status screen (done, KA, 5/2/2012)
    • Add "BURT" status indicators on the cds status screen (done, KA, 5/2/2012)
  • SCRIPT
    • Locking scripts with integrator triggers / disabling when unlocked
    • Daily diagnosis of the MC spot positions (there must be something already...)
    • Daily/occasional adjustment of the incident axis on the MC
    • Suspension "misalign/restore" script is too wild => make a new script (Done, JCD, 7May2012)
    • IFO_CONFIGURE scripts still do a burt restore, so step the optics.  Need to remove optic alignment from config .snap files, use reg restore script instead.
    • OPLEV/OSEM trending script before the IFO work for diagnosis.
    • Auto-locker for PMC/Arm/etc
  • Video
    • If each video screen has a caption, that would be great
    • GUI interface of "videoswitch" Mike!
    • Mouse Jiggler for Zita (called Swinput?)
  • Ubuntu
    • burttoday is not aliased in ubuntu. burttoday:      aliased to cd /opt/rtcds/caltech/c1/burt/autoburt/today/
    • ASS doesn't run on Ubuntu!
      ezcawrite: error while loading shared libraries: libca.so: cannot open shared object file: No such file or directory
      ezcaswitch: error while loading shared libraries: libca.so: cannot open shared object file: No such file or directory
       
  6594   Wed May 2 21:04:06 2012 DenUpdatePEMEM 172 coherence

I measured coherence between 2 EM 172 microphones in a "quiet room" with SR785

em_coh.png

High-frequency noise (>2k) is SR785 noise - I'm not using any amplifier now, the signal from microphone is sent directly to SR785 and is weak at high frequencies.

  6593   Wed May 2 19:47:20 2012 DenUpdatePEMEM 172 nonlinearities

I've checked new small EM172 microphones for nonlinearities using Koji's Mac Book speakers. EM172 + Mac nonlinearities are presented at the figure

500Hz_noise.png

The Mac's sound frequency was specified to be 500 Hz. Harmonics are seen at 1k, 1.5k, but their amplitude is ~30 times less.

  6592   Tue May 1 17:42:15 2012 Mike J.UpdateComputersSensoray

I've upgraded the Sensoray GUI so it can now switch the video channel it receives, thanks to the videoswitch script.

V4L2_Capture_Demo_r01.png

  6591   Tue May 1 08:18:50 2012 JamieUpdateCDSFrame Builder is down

Quote:

 

I tried restarting the fb in two different ways.  Neither of them re-established the connection to dtt or epics.

 Please be conscious of what components are doing what.  The problem you were experiencing was not "frame builder down".  It was "dtt not able to connect to frame builder".  Those are potentially completely different things.  If the front end status screens show that the frame builder is fine, then it's probably not the frame builder.

Also "epics" has nothing whatsoever to do with any of this.  That's a completely different set of stuff, unrelated to DTT or the frame builder.

  6590   Mon Apr 30 22:58:57 2012 JenneUpdateElectronics11MHz Marconi set to default after power outage

After a power outage, a Marconi comes back to it's defaults.  It needed to be reset to the values in elog 5530.  I'm putting a label on the Marconi so we don't have to look it up next time.

Before fixing the Marconi, POY11, AS11 and AS55 all looked like noise, no real signals, even though the arm is flashing.  Now they all look PDH-y, so things are better.

  6589   Mon Apr 30 22:41:34 2012 JenneUpdateSUSETMY oplev power supply dead

Quote:

And has the temp controller of the green been recovered too? (Push enable button)

 

 Just now, yes.

  6588   Mon Apr 30 21:24:43 2012 KojiUpdateSUSETMY oplev power supply dead

And has the temp controller of the green been recovered too? (Push enable button)

 

  6587   Mon Apr 30 21:05:32 2012 JenneUpdateSUSETMY oplev power supply dead

[Jenne, MikeJ]

The ETMY oplev quad sum was ~0, so we went to investigate.  The power supply was dead.  Keying it on and off didn't fix things, and it was certainly getting power since the front indicator light was flickering.  We replaced it with a different power supply, and the oplev is back.

Also, the Y-green AUX laser was off, presumably from the power outage a while back.  I turned it back on.

 

EDIT JD:  The first power supply we used didn't work....the oplev was on for a few minutes, then went off again.  We switched to one of the new (still JDSU) power supplies from Edmund Optics, and it's been happily working for at least an hour now.

  6586   Mon Apr 30 20:43:33 2012 SureshUpdateCDSFrame Builder is down

Quote:

Quote:

Quote:

Frame builder is down.  PRM has tripped its watch dogs.  I have reset the watch dog on PRM and turned on the OPLEV. It has damped down.  Unable to check what happened since FB is not responding.

There was an minor earthquake yesterday morning which people could feel a few blocks away.  It could have caused the the PRM to unlock.

Jamie,Rolf,  is it okay or us to restart the FB?  

 If it's down it's alway ok to restart it.  If it doesn't respond or immediately crashes again after restart then it might require some investigation, but it should always be ok to restart it.

I tried restarting the fb in two different ways.  Neither of them re-established the connection to dtt or epics.

1) I restarted the fb from the control room console with the 'shutdown' command.  No change.

2) I halted the machine with 'shutdown -h now' and restarted it with the hardware reset button on its front-panel. No change.

The console connected to the fb showed that the network file systems did not load.  Could this have resulted in failure to start several services since it could not find the files which are stored on the network file system?

The fb is otherwise healthy since I am able to ssh into it and browse the directory structure.

[Mike, Rana]

The fb is okay.  Rana found that it works on Pianosa, but not on Allegra or Rossa.  It also works on Rosalba, on which Jamie recently installed Ubuntu.

The white fields on the medm  'Status' screen for fb are an unrelated problem.

 

 

  6585   Mon Apr 30 18:46:34 2012 ranaUpdateComputersmegatron

Last week I found that megatron was still off after the power outage. Apparently, the power recovery checklist had not been followed during the recovery.

Please remember to use the checklist and post the checklist results after each power outage. Megatron is now on and functioning.

  6584   Mon Apr 30 16:56:05 2012 SureshUpdateCDSFrame Builder is down

Quote:

Quote:

Frame builder is down.  PRM has tripped its watch dogs.  I have reset the watch dog on PRM and turned on the OPLEV. It has damped down.  Unable to check what happened since FB is not responding.

There was an minor earthquake yesterday morning which people could feel a few blocks away.  It could have caused the the PRM to unlock.

Jamie,Rolf,  is it okay or us to restart the FB?  

 If it's down it's alway ok to restart it.  If it doesn't respond or immediately crashes again after restart then it might require some investigation, but it should always be ok to restart it.

I tried restarting the fb in two different ways.  Neither of them re-established the connection to dtt or epics.

1) I restarted the fb from the control room console with the 'shutdown' command.  No change.

2) I halted the machine with 'shutdown -h now' and restarted it with the hardware reset button on its front-panel. No change.

The console connected to the fb showed that the network file systems did not load.  Could this have resulted in failure to start several services since it could not find the files which are stored on the network file system?

The fb is otherwise healthy since I am able to ssh into it and browse the directory structure.

  6583   Mon Apr 30 13:58:25 2012 JamieUpdateCDSFrame Builder is down

Quote:

Frame builder is down.  PRM has tripped its watch dogs.  I have reset the watch dog on PRM and turned on the OPLEV. It has damped down.  Unable to check what happened since FB is not responding.

There was an minor earthquake yesterday morning which people could feel a few blocks away.  It could have caused the the PRM to unlock.

Jamie,Rolf,  is it okay or us to restart the FB?  

 If it's down it's alway ok to restart it.  If it doesn't respond or immediately crashes again after restart then it might require some investigation, but it should always be ok to restart it.

  6582   Mon Apr 30 13:00:50 2012 SureshUpdateCDSFrame Builder is down

Frame builder is down.  PRM has tripped its watch dogs.  I have reset the watch dog on PRM and turned on the OPLEV. It has damped down.  Unable to check what happened since FB is not responding.

There was an minor earthquake yesterday morning which people could feel a few blocks away.  It could have caused the the PRM to unlock.

Jamie,Rolf,  is it okay or us to restart the FB?  

  6581   Fri Apr 27 13:32:06 2012 DenUpdatePEMseism channels

A few weeks ago I found that GUR2_X signal is biased from 0 to 800 counts in average. I decided that the corresponding channel in the readout box is bad - adds DC voltage to the signal. I stopped using GUR2_XYZ channels of the seism readout box. Now the same thing happened with the GUR1_XYZ channels. I checked the signals coming out from the seism box with the oscilloscope and they were fine. So the problem is not in the readout box. Then I applied 1 V sine wave to the input of AA board to the GUR1_X and ACC_MC1_Z channels. GUR1_X channel still shows noise. Something is wrong with these channels inside the AA board or in the ADC.

pem.png

 

Edited by Den: GUR1_XYZ_IN1 signals are empty though GUR1_XYZ are fine. So the problem is just that GUR1_XYZ_IN1 are not acquired for now though some of the ACC_IN1 channels contain the signal. I need to correct .ini files.

  6580   Fri Apr 27 12:12:14 2012 DenUpdateCDSc1ioo is back

Rolf came to the 40m today and managed to figure out what the problem is. Reading just dmesg was not enough to solve the problem. Useful log was in

>> cat /opt/rtcds/caltech/c1/target/c1x03/c1x03epics/iocC1.log

Starting iocInit
The CA server's beacon address list was empty after initialization?
iocRun: All initialization complete
sh: iniChk.pl: command not found
Failed to load DAQ configuration file

iniChk.pl checks the .ini file of the model.

>> cat /opt/rtcds/rtscore/release/src/drv/param.c


int loadDaqConfigFile(DAQ_INFO_BLOCK *info, char *site, char *ifo, char *sys)
{

  strcpy(perlCommand, "iniChk.pl ");
  .........
  strcat(perlCommand, fname); // fname - name of the .ini file
  ..........
}

So the problem was not in the C1X03.ini. The code could not find the perl script though it was in the /opt/rtcds/caltech/c1/scripts directory. Some environment variables are not set. Rolf added /opt/rtcds/caltech/c1/scripts/ to $PATH variable and c1ioo models (x03, ioo, gcv) started successfully. He is not sure whether this is a right way or not, because other machines also do not have "scripts" directory in their PATH variable.

>> cat /opt/rtcds/caltech/c1/target/c1x03/c1x03epics/iocC1.log

Starting iocInit
The CA server's beacon address list was empty after initialization?
iocRun: All initialization complete

Total count of 'acquire=0' is 2
Total count of 'acquire=1' is 0
Total count of 'acquire=0' and 'acquire=1' is 2

Counted 0 entries of datarate=256     for a total of 0
Counted 0 entries of datarate=512     for a total of 0
Counted 0 entries of datarate=1024     for a total of 0
Counted 0 entries of datarate=2048     for a total of 0
Counted 0 entries of datarate=4096     for a total of 0
Counted 0 entries of datarate=8192     for a total of 0
Counted 0 entries of datarate=16384     for a total of 0
Counted 0 entries of datarate=32768     for a total of 0
Counted 2 entries of datarate=65536     for a total of 131072

Total data rate is 524288 bytes - OK

Total error count is 0

Rolf mentioned about automatic set up of variables - kiis of smth like that - probably that script is not working correctly. Rolf will add this problem to his list.

  6579   Fri Apr 27 09:27:48 2012 DenUpdateCDSsus watchdogs?

Quote:

Why are all the suspension watchdogs tripped?  None of the suspension models are running on c1ioo, so they should be completely unaffected.  Steve, did you find them tripped, or did you shut them off?

In either event they should be safetly turned back on.

 I've turned off the coils. Though non of them are on the c1ioo, who knows what can happen when we'll try to run the models again.

ELOG V3.1.3-