  7170   Tue Aug 14 04:37:06 2012 YoichiSummaryLSCXARM Open Loop Gain

Yoichi, Rana

Here is the open loop gain of the XARM loop.

The reference is from the pre-upgrade era. We get the extra phase delay because we have two anti-aliasing filters. One is the hardware filter at about 7kHz for 16kHz sampling. This filter should have been replaced to the one for 64kHz sampling but it has not yet happened. The second one is the software anti-aliasing filter applied when down sampling from 64kHz to 16kHz. So we have double AA filters, which are the culprits for the extra phase delay.

We should either replace the hardware AA filter to the 64kHz one (preferred way), or change the software AA filter to a less aggressive one (easier temporary fix).

Attachment 1: xarm-opltf.png
xarm-opltf.png
  7171   Tue Aug 14 04:53:45 2012 YoichiSummaryLSCX-Arm noise spectrum

Yoichi, Rana

Here is the noise spectrum of the X-arm error signal along with the TRX DC power fluctuations.

The spectra were taken while the whitening filters for POX11 were OFF.

EDIT (Integrity Fairy): Shall we assume these units are "Intergalactic translational qubits/sqrt(Hz)"?

Attachment 1: xarm-spectrum.png
xarm-spectrum.png
  7184   Tue Aug 14 22:16:46 2012 JenneUpdateLSCLSC whitening triggers

I'm ~30% of the way through implementing LSC whitening filter triggers.  I think that everything I have done should be compile-able, but please don't compile c1lsc tonight.  I haven't tested it, and some channel names have changed, so I need to fix the LSC screen when I'm not falling asleep.

Also, Rana pointed out that we may not want the whitening to trigger on immediately upon acquiring lock - if there are other modes ringing down in the cavity, or some weird transients, we don't want to amplify those signals.  We want to wait a second or so for them to die down, then turn on analog whitening.  Jamie - do you know how long the "unit delay" delays things in the RCG?  Do those do what I naively think they do?  I'll ask you in the morning.

  7187   Wed Aug 15 04:03:55 2012 ranaSummaryLSCY-Arm Locking

0) Did a bunch of alignment to get beams roughly centered on ETMY and ITMY and maximize power. Adjusted the aperture and focus on ETMY camera to get nice image. Camera needs to be screwed in tightly and cables given some real strain relief, Steve.

1) snapshots not working on many MEDM screens. Who's on top of this?

2) save/restore not working for PZT2 sliders

3) changed power and filter triggers on yarm to match xarm

4) yarm filters copied from xarm (need to handtune RGs)

5) DTT wasn't working on rossa. Used the Date/Time GUI to reset the system time to match fb and then it stopped giving 'Test Timed Out'. Jamie check rossa ntpd.

6) Removed the high 3.2 Hz RG filter. We don't have any sharp features like that in the spectrum.
   ---then added it back. The 3.2 Hz comes and goes depending on what Yoichi is doing over in the MC area. Leaving it in by default, but lowering the Q from 2 to 1.5.

7) Attached is the noise spectra, coherence, and loop gain model for this yarm condition. For the plant model, I assume a pendulum (f=1 Hz, Q = 9) and a cavity pole of 1600 Hz. Gain is scaled to set the UGF at 165 Hz (as guessed by looking at the servo gain peaking frequency). This cheezy model doesn't include any of the delays from DAC, AA, or AI. Eric and Sasha should have something more useful for us by Friday.

8) Change the DQ channels: need XARM and YARM IN1 at 16k. e.g. XARM_ERR, etc.

9) To get the DTT plots to make thumbnails in the elog, I print a .ps file and then use 'epstopdf' to make the PDF.

Attachment 1: yArmNoise_120815.png
yArmNoise_120815.png
Attachment 2: yarm.pdf
yarm.pdf
  7188   Wed Aug 15 09:09:45 2012 jamieUpdateLSCLSC whitening triggers

Quote:

I'm ~30% of the way through implementing LSC whitening filter triggers.  I think that everything I have done should be compile-able, but please don't compile c1lsc tonight.  I haven't tested it, and some channel names have changed, so I need to fix the LSC screen when I'm not falling asleep.

Also, Rana pointed out that we may not want the whitening to trigger on immediately upon acquiring lock - if there are other modes ringing down in the cavity, or some weird transients, we don't want to amplify those signals.  We want to wait a second or so for them to die down, then turn on analog whitening.  Jamie - do you know how long the "unit delay" delays things in the RCG?  Do those do what I naively think they do?  I'll ask you in the morning.

The unit delay delays for a single cycle, so I think that's not what you want.  I'm not sure that there's an existing part to add delays like that.

We also need to be a little clever about it, though, since we'll want it to flip off if we loose lock during the delay.

  7191   Wed Aug 15 11:44:35 2012 jamieSummaryLSCntp installed on all workstations

Quote:

5) DTT wasn't working on rossa. Used the Date/Time GUI to reset the system time to match fb and then it stopped giving 'Test Timed Out'. Jamie check rossa ntpd.

ntp is now installed on all the workstations.  I also added it to the /users/controls/workstation-setup.sh script

  7211   Fri Aug 17 00:16:30 2012 EricSummaryLSCYARM Calibration

I modified my Simulink model of the YARM to match the new filter modules Rana installed on YARM. I also scaled the open loop transfer function of the model to fit the measured open loop transfer function at the unity gain frequency, as shown in the figure below. From this I produced the length response function correctly scaled, also shown below.  Then I applied the calibration factor to the YARM data measured in /users/Templates/Y-Arm_120815.xml. Both the uncalibrated and calibrated spectra are included below.

 

 

Attachment 1: olg_model_meas.png
olg_model_meas.png
Attachment 2: length_response_model.png
length_response_model.png
Attachment 3: yarm_uncal_power_spec.pdf
yarm_uncal_power_spec.pdf
Attachment 4: yarm_cal_power_spec.pdf
yarm_cal_power_spec.pdf
  7213   Fri Aug 17 04:54:01 2012 Yoichi, KojiSummaryLSCPRMI Locking

 To taste the strangeness of the current 40m PRC, I locked the PRMI with the guide of Koji.

We first aligned MICH by mostly tweaking ITMX, assuming that ITMY is in a good place as the Y-arm locks. MICH lock was stable.

Then we restored the IFO to the PRM_SBres mode. With a bit of alignment work on PRM and gain tweaking, the PRMI locked.

Yes, the beam spots look UGLY !

Also the PRMI was not so stable. Especially, when the alignment fluctuates, the optical gain changes and the loop becomes temporarily unstable. We took POP_DC as the guide for the gain change and normalized the PRCL error signal with it. To do this smoothly, we first changed the input matrix to route the PRCL error signal, which is REFL33_I, so that the signal also goes to the MC filter bank. Then with dtt, we monitored the spectra of the PRCL_IN1 and MC_IN1. We tweaked the value of the element in the normalization matrix for the MC path until the two spectra look the same (at this moment, the normalizing factor for the PRCL path was still zero). During this process, we noticed that the MC path signal (normalized by POP_DC) is noisier at above 500Hz. This was because the POP_DC has a large noise at high frequencies. So we put a low pass filter (100Hz 2nd order Butterworth) to the POP_DC filter bank to reduce the noise. Then the two spectra looked almost the same. The correct normalization factor found in this way was 0.03. So we put this number in the normalization matrix for PRCL. It did not break the PRMI lock.

 

After the normalization is turned on, the PRMI lock became somewhat more stable. However, the POP_DC was still fluctuating a lot, especially when the alignment is good. So I made a boost filter: 5Hz pole Q=2, 15Hz zero Q=1.5. I also made this filter automatically triggered when the PRMI is locked. This made the PRMI lock acquisition quicker. However, still the POP_DC fluctuation is large. It seems that the alignment of PRC is really fluctuating a lot.

 The current UGF of PRMI is about 150Hz with the phase margin over 50deg.

 

 

 

 

Attachment 1: AS_1029238601.jpg
AS_1029238601.jpg
Attachment 2: POP_1029238616.jpg
POP_1029238616.jpg
Attachment 3: REFL_1029238629.jpg
REFL_1029238629.jpg
Attachment 4: PRMI-OPLG.png
PRMI-OPLG.png
  7224   Sat Aug 18 03:55:12 2012 YoichiSummaryLSCX-arm locking again

Tonight, I worked on the X-arm locking again. I did not have any significant progress, but observed several issues and will give some suggestions for future work here.

What I did tonight was basically re-alignment of the X-arm (because Rana touched the PZT mirrors for the Y-arm alignment, the X-arm alignment was screwed up). Then I measured the open loop gain. Of course it was almost identical to the one posted in this entry. It reminded myself of how small the phase bubble is. This means we have to finely adjust the gain to set the UGF at the right frequency, i.e. 100Hz. So I decided to do the signal normalization using the TRX power. Using the MC path method described here,  the appropriate normalization coefficient was determined to be 1.6, when the XARM gain is set to 0.05. Using burtgooey, I updated the burt snapshot used by the X-arm restore script.

Now I observed the following things:

When the normalization is used, the lock itself is stable, but the lock acquisition takes loner (i.e. fails more often).

I don't know the exact reason, but here is my guess: Usually, the error signal is divided by the square root of the transmitted power to widen the linear range of the PDH error signal. However, what I'm doing here is dividing the error signal with the power itself, not the sqrt. This might distort the error signal in a not-friendly-for-lock way ? I don't know.

I checked the c1lsc FE code. There seems to be the sqrt(TRX) and sqrt(TRY) signals computed in the code. However, these are not used for the normalization. 

Now, there are two requirements. When dragging the mirrors into the resonance, we want to normalize the error signal with sqrt(TRX). When the mirrors reach the resonance, the gain of the loop must be normalized by TRX. How do we smoothly connect those two states ? Someone should spend some time on this. Maybe I will work on this in Japan.  

We really need a time delay in the filter trigger

The automatic filter trigger is awesome. However, the [0^2:5^2] filter, which is an integrator, takes time to switch on and off. Every time the cavity passes by a resonance, this filter gets turned on and off slowly, giving some large transients. This transient combined with the bad coil balance of ETMX sometimes made the optical lever of ETMX crazy. This can be avoided by turning on this filter a few seconds after the power reaches the threshold. As Rana suggested, we should be able to put an arbitrary time delay to the filter trigger.

Someone should balance the coils

The coil balance of ETMX is bad and causing the above mentioned problem. I tweaked the coil balance by injecting a sinusoidal signal (10Hz) into ETMX pos and trying to minimizing the spectral peak in the optical lever signals. Of course, this is a cheesy work. Someone should put more serious effort on this.

A civilized interferometer should have an auto-alignment capability

After my alignment work, the X-arm power got to about 0.7. (This is probably because the MC transmission power has been low for the past 5 hours or so (attachment 1)).

In anyway, after the cavity locked to the TEM00 mode, the alignment has to be automatically improved by dithering. It is anachronism to sit down and click on the MEDM screen until the power gets big enough.

 

 

Attachment 1: MC_Trans.png
MC_Trans.png
  7274   Fri Aug 24 21:00:40 2012 KojiUpdateLSCX end green investigation

I checked and fixed the X end green situation. Now the X green beam is locked with TEM00.

There are various reasons it did not lock nicely.

  • The IR beam axis was changed by Yoichi and Rana (ELOG #7169). So the green axis also had to be changed.
  • The end green optics is really "BS". Anytime I see it, I feel disgusted. Because of 3D steering mirrors, cross couplings
    between yaw and pitch are big. This makes the alignment hard.
  • Even with acceptable alignment, the lock was only momentarily. I found the slow control was on. This pushed the frequency
    too much and made the lock unstable.
  • The slow control screen was broken as Jamie changed the model names but did not fix the slow screens.
    • Jamie saids (ELOG #7011): Fix the c1sc{x,y}/master/C1SC{X,Y}_GC{X,Y}_SLOW.adl screens. 
      I need to figure out a more consistent place for those screens.

Now some action items are left:

- IR TRX is not aligned.
- X end green needs precise alignment.
- PSL GR TRX is not aligned.

These will be checked on Sunday.

- End green setup is horrible. => Manasa and I should work on this together.

  7311   Wed Aug 29 19:28:41 2012 Elli KingUpdateLSCSetup for a cavity scan or the input mode cleaner

 Riju, Elli

Today we prepared our experimental setup to take a cavity scan of the input mode cleaner, which we want to measure in the next day or so.  Attached is a diagram of our setup.

What we want to do is to inject a set of sidebands into the PSL and sweep their frequency from 32-45 MHz (a range just over one fsr of the mode cleaner- vfsr=11MHz).  We will measure the power transmitted out of the MC using a photo-diode and demodulate this signal with our input signal from the Marconi.  From this we should be able to see the resonant frequencies of the carrier and the higher order modes.

One aspect we spent some time thinking about; whether we would be able to inject a signal into an EOM given the EOM and the Marconi are not perfectly impedance matched.  Based on Kiwamu’s previous e-log entries designing the EOM, we decided that injecting a signal in 32-45 MHz region at 15dBm is similar to injecting the 29.5MHz sideband (at the same power level with very similar input impedance.) Fingers crossed we don’t blow anything up first week on the job.

Attachment 1: 40m_cavity_scan_diagram.jpg
40m_cavity_scan_diagram.jpg
  7360   Fri Sep 7 12:28:09 2012 KojiUpdateLSC11&55MHz modulations turned off

11MHz modulation source was turned off (disabled) at Marconi at 12:00.

  7445   Thu Sep 27 13:05:55 2012 Eric GustafsonUpdateLSC40 meter photodiode frequency response measurement system installation

Jenne, Mike and I installed all of the post holders we could today including: REFL11, REFL33, REFL55, AS55, MCRef, POX11 and POP55.  We did not install AS110, POY or REFL165 because there are interferences that will require moving stuff around. We also did not mount POP22 because it is a peely wally ThorLabs PD that will be replaced by a strong, straight and right thinking LIGO PD in the fullness of time.  We did move it out of the way however which is no more than it deserves. Next step this afternoon Mike and I will install all of the telescopes and launching hardware.  Then with the help of Steve we will begin routing the fibers.  The splitter module will be here by next Monday, the laser by the following Friday and then we will light up the fibers. 

  7447   Thu Sep 27 16:26:11 2012 SteveUpdateLSCRF fibre protection in cable trays

Quote:

Jenne, Mike and I installed all of the post holders we could today including: REFL11, REFL33, REFL55, AS55, MCRef, POX11 and POP55.  We did not install AS110, POY or REFL165 because there are interferences that will require moving stuff around. We also did not mount POP22 because it is a peely wally ThorLabs PD that will be replaced by a strong, straight and right thinking LIGO PD in the fullness of time.  We did move it out of the way however which is no more than it deserves. Next step this afternoon Mike and I will install all of the telescopes and launching hardware.  Then with the help of Steve we will begin routing the fibers.  The 1x16 splitter module will be here by next Monday, the laser by the following Friday and then we will light up the fibers. 

 I'm proposing split loom tubing that would run in the cable tray  to protect the fibers  inside of it.  This tubing diameter in the cable tray can be 1.5-2"  and out of the tray 0.75"

 

  7448   Thu Sep 27 17:00:41 2012 Eric GustafsonUpdateLSC40 meter photodiode frequency response measurement system installation

Mike and I installed all of the telescopes and launching hardware for REFL11, REFL33, REFL55, AS55, MCRef, POX11 and POP55. On Monday afternoon Steve will work with us on the fiber routing.  Steve is buying some protective covers for the fibers.

  7513   Tue Oct 9 23:12:56 2012 JenneUpdateLSC11MHz reconnected to EOM

Riju hasn't been in the lab in a long time to do any measurements, so I put the signals back to how they should be. 

I turned off / confirmed off the things which were sending signal to the EOM:  the network analyzer, the RF generator box, and the Marconi which supplies the 11MHz. 

I removed the cavity scanning cable, and terminated it, and put the regular 11MHz cable back on the splitter.

I then turned on the RF gen box and the Marconi.  The Marconi had been off, so we were not getting any 11MHz or 55MHz out of the RF gen. box.  This is why I couldn't lock any cavities last night (duh). 

On to locking!

----------------- In other news,

While swapping out the EOM cable, I noticed that the DC power supply sitting under the POX table was supplying a weird value, 17 point something volts.  I checked on the table to remind myself why that power supply is there...it's powering an RF amplifier right after the commercial PD that is acting as POP22.  The amplifier wants +15 and GND, so I reset the power supply to 15V.  We should add this to the list of things to fix, because it's dumb.  Either we need to put in the real POP22 (long term goal), or we need to get this guy some rack power, and do the same for any amplifiers for the Beat setup.  It's a little hoakey to have power supplies littering the lab.

  7515   Wed Oct 10 02:15:14 2012 ranaUpdateLSC11 MHz reconnected to EOM

 Absolutely hokey. What are our requirements for this RFPD? What are the power levels and SNR that we want (I seem to remember that its for 22 as well as 110 MHz)? Perhaps we can test an aLIGO one if Rich has one sitting around, or if the aLIGO idea is to use a broadband PD I guess we can just keep using what we have.

  7554   Tue Oct 16 00:33:29 2012 JenneUpdateLSCPOP lens placed on POX table, 2 PRMI movies

[Evan, Jenne]

We aligned the PRMI.  We definitely can lock MICH, but we're not really sure if PRCL is really being locked or not.  I don't think it is.

Anyhow, we found 2 different places on the AS camera that we can align the PRMI.  One (middle, right hand side of the camera), we see the same weird fringing that we've been seeing for a week or two.  The other (lower left side of the camera), we see different fringing, almost reminds me more of back in the day a few months ago when the beam looked like it was expanding on each pass.  As I type, Evan is uploading the movies to youtube.  I *still* don't know how to embed youtube videos on the elog!

Also, we found both forward-going and backward-going POP beams coming out onto the POX table.  We placed the 2" lens in the path of the backwards beam, so that we can find it again.  We can't see it on an IR card, but if we put some foil where we think the beam should be, we can use a viewer to see the spot on the foil.  Poking a hole in the foil made an impromptu iris.

Youtube videos:

Lower left on camera

Middle right on camera

  7555   Tue Oct 16 02:34:38 2012 KojiUpdateLSCPOP lens placed on POX table, 2 PRMI movies

How can you lock the PRMI without the REFL beams? c.f. this entry by Kiwamu
Which signals are you using for the locking?

I think the first priority is to find the fringes of the arms and lock them with POX/POY.

As for the POP, make sure the beam is not clipped because the in-vac steering mirrors
have been supposed to be too narrow to accommodate these two beams.

  7556   Tue Oct 16 11:38:17 2012 JenneUpdateLSCMore PRMI notes from last night

Quote:

How can you lock the PRMI without the REFL beams? c.f. this entry by Kiwamu
Which signals are you using for the locking?

I think the first priority is to find the fringes of the arms and lock them with POX/POY.

As for the POP, make sure the beam is not clipped because the in-vac steering mirrors
have been supposed to be too narrow to accommodate these two beams.

I was using AS55I for PRCL, and AS55Q for MICH.  I snuck that into the last line of an unrelated elog, since I did both things at the same time: see elog 7551.  Kiwamu's measurements (elog 6283) of the PRMI sensing matrix show that the PRCL and MICH signals are almost orthogonal in AS55 (although the optickle simulation doesn't agree with that...)  He was able to lock PRMI with AS55 I&Q (elog 6293), so I thought we should be able to as well.  Locking the PRMI was supposed to help tune the alignment of the PRM, not be the end goal of the night.  Also, we only tried locking PRCL in the "middle right" configuration, not the "lower left" configuration, but we were seeing what looked like recycling flashes only in the "lower left" configuration.

I agree in principle that we should be working on the arms. However, since we can't use the old steer-the-beam-onto-the-cage trick to find the beam, I was hoping that we could steer the beam around and see some light leaking out of the ETM, onto the end table.  However, with the 1% transmission of the ITMs and ~10ppm transmission of the ETMs, there's not a lot of light back there.  I was hoping to align the PRMI so that I get flashes with a gain of 10 if I'm lucky, rather than just the 5% transmission of the PRM.  With the PRMI aligned, I was expecting:

(1W  through Faraday) * (10 PR gain) * (0.5 BS transmission) * (0.01 ITM transmission) * (10ppm ETM transmission) = 0.5uW on the ETM tables during PRCL flashes

I was hoping that things would be well enough aligned that I could just go to the end table, and see the light with a viewer, although as I type this, I realize that if the beam was not on the end table (or even if it was...) any time I move the PZTs, I'd have to completely realign the PRMI in order to see the flashes.  This seems untenable, unless there are no other options.

We then got sidetracked by trying to see the POP beam, and once we saw the POP beam we wanted to put something down so we could find it again.  POP is also small, but not as small as expected at the end:

(1W  through Faraday) * (10 PR gain) * (20ppm PR2 transmission) = 0.2mW on POP during PRCL flashes.

POP was very difficult to see, and we were only able to see it by putting the foil in the beam path, and using a viewer.  I think that we once were able to see it by looking at a card with the viewer, but it's much easier with the foil.  I'd like to find an iris that is shiny (the regular black iris wasn't helpful), to facilitate this alignment.  Since we were just looking at the reflection off of the foil, I have no comment yet about clipping vs. not clipping.  I do think however that the forward-going beam may have been easier to find....when the PRMI alignment drifted, we lost the beam, but I could still see the forward-going beam.  Probably I should switch to that one, since that's the one that was lined up with the in-vac optics. 

Summary:

Ideas are welcome, for how to align the beam to the Yarm (and later to the Xarm), since our old techniques won't work.  Aligning the PRMI was a distraction, although in hopes of getting flashes so we could see some light at the end tables.  I'm going to go see if I can look through a viewport and see the edges of the black glass aperture, which will potentially be a replacement for the steering-on-the-cage technique, but if that doesn't work, I'm running out of ideas.

  7738   Wed Nov 21 21:06:13 2012 AyakaUpdateLSCcalibration of arms

Motivation

In order to estimate whether we can see acoustic coupling in arms or not, we have to calibrate signals to phase noise.

Method

I used the same method as Yuta and Jenne did (6834).
I switched from ETM locking to ITM locking since only ITM actuators are calibrated (5583), and measured the open loop transfer function and the transfer function from ITM excitation to POX/POY error signal. Then I can estimate the calibration value H [counts/m] from POY/POX error signal to displacement.

Results

Yarm; H = 9.51 x 1011 counts/m
  OL_y.pngerr_exc_y.pngPOY_disp_y.png

Xarm; H = 6.68 x 1011 counts/m
OL_x.pngerr_exc_x.pngPOX_disp_x.png

Phase noise in arms:
XY_phase.png
blue; Xarm, green; Yarm

 

Next Step

I will calibrate the acoustic signal and see if it is reasonable that we can see the acoustic coupling signal in the arms.
But I guess it is difficult. Actually I have not seen coherence between ETM feedback signals and acoustic sounds yet. (I measured acoustic noise near POX and in PSL table.)

If I find that it is hopeless, I will create some sounds and try to measure transfer function from acoustic sound to arm cavity signals.
I am interested in how the transfer function calculated by wiener filtering is different from the measured transfer function.

 

Note

I found that we do not have enough phase margin. This is why the arm locking is not so stable.

  7739   Sat Nov 24 13:58:07 2012 ranaUpdateLSCcalibration of arms

For the loop diagnosis, its best to use the method of "IN1/IN2", rather than manipulate the close loop gain. In this way, you can directly plot the swept sine measurement from DTT as the open loop gain.

Also, for reporting calibration, we should all try to record the current settings better. Anything that may change the loop gain should be recorded along with the Bode plot and the DATA must be posted to the elog - no more of just posting plots.

We need to know, e.g.

  1. what is the power in the arms?
  2. are the LSC whitening filters on?
  3. are the SUS dewhitening filters on?
  4. What normalization is being used in the LSC?
  5. What digital filters are on in the X/YARM loop filter bank?

Resistance is feudal.

  7743   Mon Nov 26 10:42:06 2012 AyakaUpdateLSCcalibration of arms

I uploaded a zip file that contains data files used for the calibration.

OLTF_x/y.txt: the open loop transfer function (measured by IN1/IN2 in arm servo filter bank).
coh_x/y.txt: coherence of OLTF. I used the data where coherence > 0.98.
ext_err_x/y.txt: the transfer function from ITM excitation signal to POX/POY error signal.
coh_x2/y2.txt: coherence of ext_err. I used the data where coherence > 0.98.

The LSC whitening filter was off because the xarm was unlocked when the POX Q whitening filter was turned on. (We have to study what was wrong.)
The SUS whitening filters were on.
The all digital filters except +6dB filter were on.

Attachment 1: armcalib.zip
  7800   Sat Dec 8 04:12:38 2012 DenUpdateLSCprcl

 Today I wanted to check that AS and REFL beams are real and contain proper information about interferometer. For this I locked YARM using AS55_I and REFL11_I. Then I compared spectrum with POY11_I locking. Everything is the same. I've also adjusted phase rotations of AS55 (0.2 ->24) and REFL11 (-34.150 -> -43).

Then I've locked MICH and aligned EMTs such that ASDC was close to zero. Then I locked PRCL and aligned PRM. Power buildup was 50. 

IMG_0118.JPG

  7801   Sun Dec 9 01:21:51 2012 DenUpdateLSCbeam inside DRMI is clipping on PR3 Tip-Tilt

Quote:

 Today I wanted to check that AS and REFL beams are real and contain proper information about interferometer. For this I locked YARM using AS55_I and REFL11_I. Then I compared spectrum with POY11_I locking. Everything is the same. I've also adjusted phase rotations of AS55 (0.2 ->24) and REFL11 (-34.150 -> -43).

 I studied more carefully beam path inside DRMI using PRM face camera and found that beam is clipping on PR3 edge.

Step 1: PRCL LOCK, MICH LOCK, power build up 30.

Note: left is right and vice versa on the PRM camera

prcl_lock.mjpg

 Step 2: PRLC - UNLOCK, MICH - LOCK, PRM is still aligned. Right photo is AS port. I've slightly misaligned ITMs such that disturbance of AS beam is clearly seen.

PRM_UNLOCK.bmp       AS_UNLOCK.bmp

 

Step 3: PRCL - UNLOCK, MICH - LOCK, PRM misalined in yaw such such that the beam LASER -> PRM -> PR2 -> PR3 -> BS -> ITMX -> BS -> PR3 -> PR2 -> PRM -> PR2 -> PR3 is completely clipped on the TT edge. AS beam is now not clipped.

PRM_MISALIGN.bmp    AS_MISALIGN.bmp

So the conclusion is that when PRC is not locked and beam is thin, it can avoid clipping. When PRC locked, beam size grows and it starts to clip. I think we need to move the mount next to PR3 because of it we to not have enough space to align the TT.

Step 4: PSL shutter is closed.

PRM_BLOCK.bmp

  7802   Sun Dec 9 17:51:53 2012 ranaUpdateLSCbeam inside DRMI is clipping on PR3 Tip-Tilt

 

 Some explanation of how you define power buildup please. Also some plots showing the evidence.

  7803   Mon Dec 10 03:02:03 2012 DenUpdateLSCbeam inside DRMI is clipping on PR3 Tip-Tilt

Quote:

 

 Some explanation of how you define power buildup please. Also some plots showing the evidence.

 I think about power buildup as a ratio of the power in the cavity when it is locked and unlocked = (POYDC_LOCKED - POYDC_OFFSET) / (POYDC_UNLOCKED - POYDC_OFFSET). I do not multiply this number by PRM transmission.

POYDC_OFFSET = -0.006

POYDC_UNLOCK = 0.063

For example, on the plot below power buildup is 15.

PRCL_LOCK.png

  7804   Mon Dec 10 10:13:41 2012 DenUpdateLSCbeam inside DRMI is clipping on PR3 Tip-Tilt

 

That's OK, but its best to use standard notation. The power recycling gain is defined as the power incident on the BS divided by the power incident on the PRM from the laser side. You should also compare it with the PRC gain that you expect from mirror transmissions.

  7806   Mon Dec 10 22:34:34 2012 DenUpdateLSCbeam inside DRMI is clipping on PR3 Tip-Tilt

Quote:

 

That's OK, but its best to use standard notation. The power recycling gain is defined as the power incident on the BS divided by the power incident on the PRM from the laser side. You should also compare it with the PRC gain that you expect from mirror transmissions.

 I've made snapshots of PR2, PRM, ITMY and ITMX mirrors. Power buildup recycling gain (POWER BS / POWER PRM) was equal to 3-4.

PR2.bmp    PRM_LOCK.bmp    ITMY.png    ITMX.png

  7808   Tue Dec 11 09:31:47 2012 AyakaUpdateLSCOplev update for improving sensitivity

 Motivation

We observed that oplev servos affect the arm spectra badly (elog #7798). Some of them are fixed, but still they inject noise into the arms.
So I tried to turn the oplevs off and to see the acoustic noise effect. However, the mirrors moves so much that the signal does not seem to be linear any more, and the noise spectrum of arms changes especially around 60 - 100 Hz as you can see the spectrogram of YARM error signal below. This makes it difficult to find acoustic coupling noise. Therefore, I tried to fix the oplev servos so that the noise spectra do not get worse when the oplev servos are on.
POYspectrogram_nonoise.png

Checking oplev UGFs

I checked the oplev open loop transfer functions. The UGFs of oplevs are all around 1-3Hz and phase margin looks enough except the BS oplev.
The gain of the BS oplev OLTF has so low that the signal is not fed back. Moreover, there is much phase delay in the BS feedback loop than the others'.
The counts of BS oplev sum is not changed so much for this 4 months, so the oplev beam seems to hit the BS correctly.
I am not sure what makes difference.
ETMYoplev.pdfOL_BSoplevPIT.pdf

 BSoplev.pdf

 

Clipped oplev beam fixed

Den and I found the output beam of ETMY oplev was clipped the other day. Also I found the scattered beam of ITMY oplev was on the edge of the mirror inside the vacuum and it made more scattered lights.
ITMYoplev_before.jpg(before) -> IMG_0128.JPG(after)

  I fixed both of the clipped beam. But still the oplev feedback inject the noise into the arm. (red: oplev off, blue: oplev on)
   POYspe_oplev.pdf

  7812   Tue Dec 11 21:53:37 2012 AyakaUpdateLSCOplev update for improving sensitivity

[Rana, Ayaka]

The BS oplev pitch feedback came back.

OL_BSpit.pdf

The problem was that 300^2:0 filter was off. And I turned on all the low pass filters (ELP35), then the oplev servo does not seem to inject big noise into the arms as long as I see the spectra of POY and POX. These low-pass filters will be modified tomorrow so that the acoustic coupling noise is minimized.

BSoplevservo.png

  7814   Wed Dec 12 11:49:05 2012 AyakaUpdateLSCAcoustic noise in POX and AS error signal

[Koji, Ayaka]

Last night, I injected acoustic noise at POX table and AS table with oplev controls on (LPF is on).

 

1. acoustic noise at the POX table

I set the microphones and speakers at the POX table and see the acoustic coupling.

POX_noise4.pdf

I could see slight change around 40 Hz. This can be caused by the oplev feedback loop because the speaker was on the same table as the ITMX oplev.

 

2. acoustic noise at the AS table

I controlled XARM with AS error signal and set the microphones and speaker on the AS table.

AScontXARM_acoustic.pdf

The resonance a 200 Hz seemed to be enhanced. But still we are not sure that it is caused by acoustic noise. Because this resonance is enhanced when the OL gain is high, and the gain adjustment was so critical that this resonance was easily enhanced even when the acoustic noise is not injected. And sometimes it has gone away.

  7820   Thu Dec 13 03:20:48 2012 DenUpdateLSCbeam inside DRMI is clipping on PR3 Tip-Tilt

Quote:

 

 I've made snapshots of PR2, PRM, ITMY and ITMX mirrors. Power buildup recycling gain (POWER BS / POWER PRM) was equal to 3-4.

           

 We've looked at PR2 face camera when PRM, BS and one of the ITMs were aligned. We saw an extra beam at PR2 when ITMX was aligned (right plot). This spot stays on the PR2 when prcl is locked.

PR2_ITMX.png   PR2_ITMY.png

Then we looked at PR3 transmission mirror and saw that the main beam is not on the edge of the mirror. Secondary beam is clipping on the mirror mount of PR3 that we see on BS_PRM camera.

PR3_LOCK.png

Measured beam spot positions:

Optics Pitch, mm Yaw, mm
ITMX 5.6 1.5
ETMX -1.5 1.5
ITMY 4.8 -1.5
ETMY -1.4 5.6
PRM 2.7 4.1

"+" for pitch means that the beam is too high, "-" too low

"+" for yaw means that the beam is left if you look from the back, "-" is right

Beam spots were measured using x, y arm and prcl locking to the carrier.

  7829   Fri Dec 14 03:32:51 2012 AyakaUpdateLSCHigh frequency noise in AS signal

I calibrated the AS error signal into the displacement of the YARM cavity in the same way as I did before (elog).

The open loop transfer function is:

XARM_OL.pdf


The transfer function from ITMX excitation to AS error signal is:

AS_ITMXexc.pdf

Then I have got the calibration value : 5.08e+11 [counts/m]

The calibrated spectrum in unit of m/rtHz is

ASspe_noise.pdf

REF0: arm displacement
REF1: dark noise + demodulation circuit noise + WT filter noise + ADC noise (PSL shutter on)
REF2: demodulation circuit noise + WT filter noise + ADC noise (PD input of the circuit (at 1Y2) is connected to the 50 Ohm terminator)
(The circuit and WT filter seem to be connected at back side of the rack. Actually there is a connector labelled 'I MON' but it is not related to C1:LSC-ASS55_I_ERR)

Also we changed the AS gain so that ADC noise does not affect:

ASgain.png

However, this did not make big change in sensitivity. I guess this means that circuit noise limits the sensitivity at higher frequencies than 400 Hz.
I tried to adjust the AS gain carefully but I could not do that because of the earthquake. Further investigation is needed.

 

Attachment 5: ASspe.tar.gz
  7832   Fri Dec 14 09:31:59 2012 ranaUpdateLSCHigh frequency noise in AS signal

This is NOT calibrated. Its sort of calibrated in the 500-1000 Hz area, but does not correctly use the loop TF or the cavity pole.

As for the noise, remember that the whole point of changing the AS whitening gain was to turn on the whitening filter AFTER locking. With the WF OFF, there's no way that you can surpass the ADC noise limit.

Quote:

I calibrated the AS error signal into the displacement of the YARM cavity in the same way as I did before (elog).

  7833   Fri Dec 14 10:09:30 2012 AyakaUpdateLSCHigh frequency noise in AS signal

Quote:

This is NOT calibrated. Its sort of calibrated in the 500-1000 Hz area, but does not correctly use the loop TF or the cavity pole.

As for the noise, remember that the whole point of changing the AS whitening gain was to turn on the whitening filter AFTER locking. With the WF OFF, there's no way that you can surpass the ADC noise limit.

Quote:

I calibrated the AS error signal into the displacement of the YARM cavity in the same way as I did before (elog).

No, I did not apply open loop TF to it (actually I could not measure the open loop TF because of the earthquake last night). So I should not have said it was the displacement.

Also I changed the AS gain with whitening filter on and xarm locked. Still it does not make any change.

  7835   Fri Dec 14 16:35:38 2012 AyakaUpdateLSCHigh frequency sensitivity improved

Since I found that the the AS sensitivity seems to be limited by circuit noise, I inserted a RF amplifier just after the AS RF output.
Now, the sensitivity is improved and limited by the dark noise of the PD.

ASspe_noise_amp.pdf

(Note: I did not apply the open loop TF on this xml file.)
REF3: dark noise + circuit noise + WT filter noise + ADC noise
REF4: circuit noise + WT filter noise + ADC

With this situation, I injected the acoustic noise:

ASspe_acoustic_amp.pdf

REF5, 6, 7: with acoustic excitation
no reference: without acoustic excitation

We could see the coherence only at the same frequencies, around 200 Hz as we saw before (elog).

Attachment 3: ASnoise.tar.gz
  8053   Sun Feb 10 18:00:13 2013 yutaSummaryLSCPR2-flipped half-PRC spectra/OLTF

To compare with future PRMI locking, I measured spectra of POPDC and feedback signal. I also measured openloop transfer function of half-PRC locking.
Beam spot motion was at ~ 2.4 Hz, not 3.3 Hz.

Results:
  Below is uncalibrated spectra of POPDC and LSC feedback signal (C1:LSC-PRM_OUT).
POPDCLSCPRM.png

  Below is openloop transfer function of the half-PRC locking loop. UGF is ~ 120 Hz and phase margin is ~ 45 deg. This agrees with the expected curve.
LSCPRCLOLTF.png

  Data was taken when half-PRC was locked using REFL11_I as error signal and actuating on PRM.


Discussion:

  For comparison, POPDC when PRMI was locked in July 2012: elog #6954 and PRCL openloop transfer function: elog #6950.

  Peak in the spectra of POPDC and feedback signal was at ~ 3.3 Hz in July 2012 PRMI, but it is now at ~ 2.4 Hz in half-PRC. The peak also got broader.
  Is it because of the change in the resonant frequency of the BS-PRM stack? How much the load on BS-PRM changed?
  Or is it because of the change in the resonant frequency of PR2/PR3?

  Phase margin is less now because of gain boost ~ 5 Hz and resonant gain at 24 Hz.
 

  8054   Mon Feb 11 12:49:54 2013 JenneSummaryLSCResonant freq change - why? (and passive TT mode freqs)

Quote:

  Is it because of the change in the resonant frequency of the BS-PRM stack? How much the load on BS-PRM changed?
  Or is it because of the change in the resonant frequency of PR2/PR3

I claim that neither of those things is plausible.  We took out 1 PZT, and put in 1 active TT onto the BS table.  There is no way the resonant frequency changed by an appreciable amount due to that switch.

I don't think that it is the resonant frequency of the TTs either.  Here, I collate the data that we have on the resonant frequencies of our tip tilts.  It appears that in elog 3425 I recorded results for TTs 2 and 3, but in elog 3447 I just noted that the measurements had been done, and never put them into the elog.  Ooops.

Resonant frequency and Q of modes of passive tip tilts. 

  Vertical Yaw Pos Side
TT1 f0=20, Q=18 f0=1.89, Q=3.8 f0=1.85, Q=2 f0=1.75, Q=3.2
TT2 f0=24, Q=7.8 f0=1.89, Q=2.2 f0=1.75, no Q meas f0=1.8, Q=4.5
TT3 f0=20, Q=34 f0=1.96, Q="low" f0=1.72, Q=3.3 f0=1.85, Q=6
TT4 f0=21, Q=14 f0=1.88, Q=2.3 f0=1.72, Q=1.4 f0=1.85, Q=1.9
TT5 f0=20, Q=22.7 no measurement f0=1.79, Q=1.8 f0=1.78, Q=3.5

Notes:  "Serial Number" of TTs here is based on the SN of the top suspension point block.  This does not give info about which TT is where.  Pitch modes were all too low of Q to be measured, although we tried.

Tip tilt mode measurements were taken with a HeNe and PD shadow sensor setup - the TT's optic holder ring was partially obscuring the beam.

  8095   Sat Feb 16 19:23:17 2013 yutaUpdateLSCPR2 flipped PRMI locked

It is my pleasure to announce that the first lock of PR2 flipped PRMI was succeeded.



POP looks very nice. TEM00 and not wobbling.
We need more I/Q phase and gain/filter adjustment and characterization soon.

Some more details:
  MICH error signal: AS55_Q_ERR (using POP55 PD; phase rotation angle 70 deg)
  PRCL error signal: REFL11_I_ERR (phase rotation angle 80 deg)
  MICH feedback: BS (MICH_GAIN = -60)
  PRCL feedback: PRM (PRCL_GAIN = -0.5)

  8143   Sat Feb 23 07:14:58 2013 yutaUpdateLSCcan't lock Y arm

I tried to align and lock Y arm for the first time after pumping.
But I couldn't lock Y arm for more than ~1 sec. Why?


What I did:
  1. Centered IPANG/IPPOS using input TT1/TT2.

  2. Restored ITMY/ETMY slider values when it was aligned before pumping. I saw tiny flashes in TRY PD at this point.

  3. Replaced Ygreen REFL camera with ETMYT camera to see transmitted beam mode.

  4. Used TT1/TT2 and ITMY/ETMY to get ~ 0.4 peak in normalized TRY PD output (C1:LSC_TRY_OUT).

  5. Centered POY beam on POY11 PD.

  6. Changed I/Q mixing angle (C1:LSC-POY11_PHASE_R) from -61 deg to -16 deg to get good PDH signal in I_ERR (attached).

  7. Ran LSCoffsets script (now on LSC_OVERVIEW screen) to adjust PD offsets.

  8. Tried to lock Yarm with different gains, but failed. When lock is acquired, TRY fluctuates ~50 % and unlocks suddenly.


What I found:
  1. There was some OFFSETs left turned on in suspension screens. Don't leave them on!! They change alignment of the optics. I will leave it on until we complete Yarm alignment.

  2. C1:SUS-(ITMY|ETMY)_ASC(PIT|YAW) was kept oscillating the optic since Dec 17, 2013. I think this is from interrupted ASS script. Your script should restore everything when interrupted!


Next:
  - Beamspot on ITMY looks off-centered. Maybe A2L is causing unstable lock?
  - Maybe F2A is causing unstable lock?
  - More alignment?
  - FSS related? crontab related?

Attachment 1: TRYPOY.png
TRYPOY.png
  8145   Sat Feb 23 14:52:03 2013 JenneUpdateLSCETMYT camera back to normal

Quote:

 3. Replaced Ygreen REFL camera with ETMYT camera to see transmitted beam mode.

The camera that Yuta means in his elog from last night/this morning is the scattering camera at the Yend.  The reason (I think) that he had to do this is that Manasa and Jan took the cable for the ETMYT camera, and were using it for their scattering camera.  They mention in elog 8072 that they installed a camera, but they didn't say anything about having taken the ETMYT cable.  This is the kind of thing that is useful to elog!

Anyhow, I have removed the Watec that belongs with the scattering setup, that Yuta borrowed, and put it back on the scattering table-on-a-pedestal. I then realigned the usual ETMYT camera (that Yuta moved out of the way to install the borrowed Watec), and put the ETMYT cable back to its usual place, connected to the Sony camera's box on the floor.

tl;dr: ETMYT camera is back to original state.

EDIT later:  I put the Watec back, since it is more sensitive to IR, so now we have a Watec in the regular ETMYT place.

  8149   Sat Feb 23 16:54:24 2013 JenneUpdateLSCcan't lock Y arm

I'm not sure that Yuta had found the real Yarm flashes last night.  When I came in today, the Yarm would flash.  However I found that the SRM was aligned, and if I misaligned it, the Yarm flashes would disappear.  So somehow the beam getting into the cavity was related to the reflection off of the SRM.

Later, I moved the TTs, leaving ITMY and ETMY alone, after having misaligned SRM (and ITMX) and I found flashes.  This wasn't ideal though, since the beam was much, much too high on IPANG (beam was half falling off the top of the lens, although yaw was pretty good).  That was also when I changed out the ETMYT camera the first time around.  The flashes on the new camera were visible, but much dimmer than with the Watec.

I tried locking the Yarm in this state, but I could never achieve a lock, even momentarily.  It almost seemed like I wasn't sending actuation signal out to the coils, although signal appeared all the way through the chain until the LSC signal summed with the local damping signal.  I also switched the LSC output matrix to try actuating on the ITM, but I was also not able to lock then.  I have switched it back to have Yarm actuate on ETMY.  I could see a nice PDH signal on POY, and nice flashes on TRY, but no lock at all.  The trigger was triggering, but still no catching of the lock.  I'm not really sure what's up.

After playing with a non-locking, poorly aligned Yarm, I started over by recentering the beam on IPPOS and IPANG using the TTs, but have not been able to get flashing in the cavity again.  After much fitzing around, I put the Watec back at ETMYT, in hopes that we can see flashes again at some point, since it's more sensitive than the old Sony.  Still no flashes though.

I have to leave, but Yuta and Manasa are here, and so I'm leaving the IFO in their custody.

  8150   Sat Feb 23 17:14:59 2013 yutaUpdateLSCcan't lock Y arm

Jenne found that;
  0. If all mirrors are "aligned," Yarm flashes.
  1. If SRM is misaligned, Yarm doesn't flash.
  2. If BS is misaligned, Yarm doesn't flash.
  3. If ITMX is misaligned, Yarm still flashes.

So, my hypothesis from this is that I was playing with " TT1 -> TT2 -> ITMY -> BS -> SRM -> BS -> Yarm "  configuration last night.
This hypothesis can explain;
  1. Why I could not get TRY peak more than 0.5 (additional BS reflection makes incident power to Yarm less).
  2. Why I had to change POY11 I/Q mixing angle by ~ 45 deg (because EOM to Yarm length changed).
  3. Why I couldn't lock Yarm stably (additional reflection by BS and SRM made too much beam jitter?).

We are now trying to get "real" Yarm flash.

  8183   Wed Feb 27 14:39:59 2013 AnnalisaUpdateLSCFibre laid for RFPD audio

 [Annalisa, Jenne, Rana, Steve]

We installed the fibres on cable trays the 1Y2 and the Control Room.

Still to do: find a power supply for the Fiboxes and plug everything in.

  8184   Wed Feb 27 14:53:02 2013 SteveUpdateLSCwhat is the Fibox ?

Fibox FBAI-M 20bit units were connected with multimode fibre.  This pair of fiber is not protected in the cable tray.

  8192   Wed Feb 27 20:50:41 2013 ManasaUpdateLSC22/110MHz path for POP

[Yuta, Manasa]

Modified POP path.

1. Removed temporary POP DC and the BS 50 (elog)
2. Introduced a 95% BS after the POP steering mirrors (95% of the signal goes to PD10CF used for POP22 and 5% goes to POP camera)
3. Output of PD10CF goes to the LSC rack through POP110 heliax cable.
4. The PD output at the LSC rack  goes through a DC block to separate DC from RF.

POP.png

We could not find a power supply slot for the amplifiers on the LSC rack. We had to put a temporary power supply in contradiction to our 'no temporary power supply' policy.

  8196   Thu Feb 28 02:43:49 2013 yutaUpdateLSCsome qualitative evidence of PRMI sideband lock

[Manasa, Yuta]

Since we have setup POP22 PD now(elog #8192), we could confirm that sideband power builds up when PRMI is sideband locked.

Plot:
  Here's some plot of PRC intra-cavity powers and MICH,PRCL error signals. As you can see from POP22, we locked at the peak of 11MHz sideband. There was oscillation at ~500 Hz, but we couldn't optimize the gain yet.
PRMIsideband.png


Movie:
  Here's 30 sec movie of AS, POP, REFL when acquiring (and losing) PRMI sideband lock. It was pretty hard to take a movie because it locks pretty seldom (~1 lock / 10 min).



Locking details:
  For MICH lock, we used ITMs instead of BS for reducing coupling between PRCL.
  Also, AS55 phase rotation angle was coarsely optimized by minimizing MICH signal in I.
  For PRCL lock, we used REFL55_I_ERR instead of REFL33_I_ERR. It had better PDH signal and we coarsely optimized phase rotation angle by minimizing PRCL PDH signal in Q.

 == PRMI sideband ==
  MICH: AS55_Q_ERR, AS55_PHASE_R = -12 deg,  MICH_GAIN = -0.1, feedback to ITMX(-1),ITMY(+1)
  PRCL: REFL55_I_ERR, REFL55_PHASE_R = 70 deg, PRCL_GAIN = -15, feedback to PRM

  We set POP22_PHASE_R = -170 deg by minimizing Q.

Issues:
 - We tried to use REFL55_Q_ERR to lock MICH, but couldn't. It looks like REFL error signals are bad.
 - We tried to use POP22_I_ERR to trigger PRCL lock, but it didn't work.

  8197   Thu Feb 28 03:25:27 2013 yutaUpdateLSCPR gain ~ 25 from PRMI carrier lock

[Manasa, Yuta]

We locked PRMI in carrier. Measured power recycling gain was ~25.

Plot:

  Here's some plot of PRC intra-cavity powers and MICH,PRCL error signals. As you can see from POPDC, cavity buildup was about 400, which means power recycling gain was ~25. Power recyling gain is fluctuating up to ~45 during lock. We need some gain normalization or something.
PRMIcarrier.png


Movie:

  Here's 30 sec movie of AS, POP, REFL when acquiring PRMI carrier lock. Although there's oscillation when acquiring lock, beam spot motion is less and stable compared with the past(before flipping PR2).



Locking details:
 == PRMI carrier ==
  MICH: AS55_Q_ERR, AS55_PHASE_R = -12 deg,  MICH_GAIN = -0.1, feedback to ITMX(-1),ITMY(+1)
  PRCL: REFL55_I_ERR, REFL55_PHASE_R = 70 deg, PRCL_GAIN = 5, feedback to PRM


Next:
  - Better filters and gains for stable lock
  - Kakeru method to measure g-factor (see elog around #1434)
  - OSA to measure g-factor

  8198   Thu Feb 28 03:41:31 2013 KojiUpdateLSCPR gain ~ 25 from PRMI carrier lock

VERY GOOD!
This is how the carrier lock PRMI should look like.

- There is more room to improve the differential ITM alignment to make the dark port more dark, then you will gain more PRG

- The AS spot is definitely clipped.