40m QIL Cryo_Lab CTN SUS_Lab CAML OMC_Lab CRIME_Lab FEA ENG_Labs OptContFac Mariner WBEEShop
  40m Log, Page 162 of 357  Not logged in ELOG logo
ID Date Authorup Type Category Subject
  9212   Mon Oct 7 10:51:18 2013 SteveUpdateSUSReplaced the laser for Optical Lever of ETMY

 Just plot.

RA: I'm not sure how to interperet this; I think that the SUM channel is divided by the SUM so that this is supposed to be RIN, but not sure. Can someone please take a look into the SUS model and then explain in the elog what the SUM normalization algorithm is?

  9220   Tue Oct 8 08:58:16 2013 SteveUpdateSUSOplev power spectrums

Quote:

 Just plot.

RA: I'm not sure how to interperet this; I think that the SUM channel is divided by the SUM so that this is supposed to be RIN, but not sure. Can someone please take a look into the SUS model and then explain in the elog what the SUM normalization algorithm is?

 

PRM is dark. PRM and SRM oplev servos are off. ETMY is not centered.

  9221   Tue Oct 8 11:29:30 2013 SteveUpdatePEMno Janitor day today

This was the second week without wet mopping the floor and wet wipe down of the vacuum envelope.

We changed filters on the 40m "HEPA" vacuum pump. It still has very dirty exhaust. Do not use it in clean room!

  9223   Wed Oct 9 11:48:10 2013 SteveUpdateVACRGA scan at day 65

 

 Valve configuration: Vacuum Normal

  9224   Wed Oct 9 15:00:48 2013 SteveUpdatePEMfreshmen tour

 [Alan, Koji, Manasa]

Record number of 50 freshmen were given tour of the 40m this afternoon.

  9229   Thu Oct 10 08:25:07 2013 SteveUpdatesafetyChris received safety training

Chris Couste, our new undergrad help received 40m specific, basic safety training yesterday.

  9230   Thu Oct 10 11:30:15 2013 SteveUpdateSUSOplev power spectrums: PIT-YAW

 

Power spetrum of ETMX and ETMY in the view of oplev  pitch and yaw.

  9241   Tue Oct 15 15:25:13 2013 SteveUpdateSUSOplev power spectrums bandwidths

  Atm1 bandwidth 0.01 Hz ETMY plot is different from

Atm2 bandwidth 0.1 Hz  normal ???????    They should be the same !!!!

  9246   Wed Oct 16 10:32:40 2013 SteveUpdateSUSFlowbench effect on oplev error signals

 ETMX _OPLEV_...ERROR signals are effected by turning ON the  flow bench motor  at the south end 

 The broadband noise [~ 5-60 Hz] is much higher when the motor is running.

The flow bench was turned off.

 

 

  9253   Fri Oct 18 08:12:08 2013 SteveUpdatePEM1000 days trends
  9254   Fri Oct 18 15:18:49 2013 SteveUpdatePEM particle count peak or ...

Sunday evening fireworks. Manasa heard it.

  9255   Mon Oct 21 09:46:12 2013 SteveUpdatePSLLaser just turned on

I have just turned on the PSL Innolight laser. The laser shut down  with unknown reason a day ago.

  9257   Tue Oct 22 10:12:33 2013 SteveUpdatePEM south arm AC intake duct checked

  The 4th week of no wet mopping of the floor  and no wet wiping  the vacuum envelope.

 

We cleaned the intakes of the south arm IFO air condition. The bottom duct have quite a bit accumulation Atm1

See wet wiped contrast on Atm2

We found 3 holes around pipes ( coming from CES ) on the east wall that has to be sealed.

After closer examination of these holes, they are sealed off well.

  9262   Wed Oct 23 09:17:33 2013 SteveUpdateSUSETMY sensors compared to ETMX

 ETMY sensors are glitching or getting kicked up.

Atm3, There is no seismic activity here.

 

  9275   Thu Oct 24 08:04:54 2013 SteveUpdateLSCPRMI + 2 ALS arms

 

 Nice work. Congratulation

  9277   Thu Oct 24 10:31:25 2013 SteveUpdateSUSETMY sensors compared to ETMX

Atm1,  The strong glitches are back.

Atm2,  SUS-ETMX & Y_SENSOR_LL Damping OFF at (ref0 & ref1). damping ON (red & blue)

ETMY sus looks OK

 

  9281   Thu Oct 24 17:18:44 2013 SteveUpdateSUSETMY oplev error signals are still bad

Quote:

 Just plot.

RA: I'm not sure how to interperet this; I think that the SUM channel is divided by the SUM so that this is supposed to be RIN, but not sure. Can someone please take a look into the SUS model and then explain in the elog what the SUM normalization algorithm is?

 

  9291   Fri Oct 25 10:45:16 2013 SteveUpdatePSLlaser drift monitor set up

Quote:

Quote:

I wonder what's drifting between the laser and the PMC? And why is it getting worse lately?

 The PMC refl is bad in pitch today, and the transmission is only 0.76, rather than our usual 0.83ish.

I did a quick, rough tweak-up of the alignment, and now we're at 0.825 in transmission.

 The PMC transmission continuously degrades. In order to see what is really drifting the laser output after PBS was sampled as shown.

  9319   Wed Oct 30 14:58:44 2013 SteveHowToGeneralshutting down a computer

 

 We have tried to ssh into c1iscey yesterday morning. It just did not work. We have just tried it again (now) and it did work.

Lesson learned: always shut down the computer from a TERMINAL Do not turn it off by the manual power switch.

  9325   Fri Nov 1 09:45:32 2013 SteveUpdateIOObeam dumps to be find

Quote:

Quote:

Steve, can you please put 2 mounted  razor dumps near the WFS for this purpose??    

            Tuesday: Razor dumps are waiting for you.

 I couldn't find any dumps near the WFS. Koji looked. I looked twice. Maybe they are spooky and absorbing all of the light?

The MC alignment was bad and the WFS were making it drift. Koji aligned the beam into the PMC. I then restored the MC suspensions to where they were 8 days ago (back when the transmission and reflection were good). With the WFS OFF, this gave us a MC trans ~ 16000. With WFS ON it goes to 17500 which is about as good as its been over the last 80 days.

I centered the beam on the WFS with the MC unlocked and also centered the beam on the whole WFS path (it was near clipping between WFS 1 & 2). Also for some reason that beamsplitter which steers the beam onto WFS1 is a R=33% (!? why is this not a R=50% ??).

Steve, please swap this out to a BS1-1064-50-1025-45S if we have one sitting around. If not, we want to add this to the CVI purchase list, but not buy until we get a bigger list together.

I also centered this newly aligned beam into the IMC onto the PSL QPDs. We should now use these as a pointing reference for the beam into the IMC.

While doing this I noticed that the beam was almost clipping on the Uniblitz shutter used to block the PSL beam. That shutter is mounted too short and was also not centered horizontally. I removed it for now so that Steve can find a more adjustable mount for it and put it back into play. The beam going into the IMC is BIG, so you have to very careful when centering the shutter. Might be that we cannot leave it at 45 deg and still get a big enough aperture.

Note #3 for Steve: please also replace the mount for last steering mirror into the IMC with a Polanski or a Superman, that black Ultima is no good. Also the dogs must be steel - no aluminum dogs for our sensitive places.

No wonder they could not find the beam dumps. Last night was Haloween. They should of just said: Trick or treat! where are the beam dumps?

  9333   Mon Nov 4 09:03:21 2013 SteveUpdatePSLPMC auto locker

Quote:

 I was working on the electronics bench and what sounded like a huge truck rolled by outside. I didn't notice anything until now, but It looks like something became misaligned when the truck passed by (~6:45-6:50 pm). I can hear a lot of noise coming out of the control room speakers and pretty much all of the IOO plots on the wall have sharp discontinuities.

I haven't been moving around much for the past 2 hours so I don't think it was me, but I thought it was worth noting.

 The PMC auto locker is not set to acquire error message made me lock PMC manually.  Arms  locked instantly: TRY 0.9 V and TRX 0.65 V

  9334   Mon Nov 4 11:37:12 2013 SteveUpdateIOOPSL output shutter installed

 

 The PSL shutter is reinstalled.The base plate is delrin for isolation and the mount height is adjustable. The last steering mirror mount to be swapped in is ready. It is sitting on the top of the ITMX optical table cover with SS dogs.

There are two reflected spots on the north side of the Uniblitz shutter. They are coming from the vacuum window. They should be trapped also.

  9343   Tue Nov 5 08:44:21 2013 SteveUpdateIOOafter last steering mirror mount swap

 The IOO Angle and IOO Position qpds  were recentered after this entry.

 Suggested corrections in elog entry #9323 are completed:

 1,  last steering mirror mount replaced by Polanski mount

 2,  PSL output shutter mount reconfigured

 IOO qpds are not centered. I failed to connect laptops to 40MARSian network.

  9347   Tue Nov 5 17:12:34 2013 SteveUpdateLSCcentered qpds

 Full list tomorrow: IP-Ang & Pos, ETMY-T, ETMY-Oplev, ETMX-T, IOO-Ang & Pos

 RA: No one in the control room this evening can understand what this ELOG means. Please use more words.

  9352   Wed Nov 6 08:33:53 2013 SteveUpdateIOOPoiting changes of PSL

Quote:

 Full list tomorrow: IP-Ang & Pos, ETMY-T, ETMY-Oplev, ETMX-T, IOO-Ang & Pos

 RA: No one in the control room this evening can understand what this ELOG means. Please use more words.

 Yesterday the last steering mirror mount on the PSL was changed, Manasa recovered the MC alignment and Jenne locked the arms.

 I centered the following qpds:  ASC-IBQPD, LSC-TRY, SUS-ETMY_OPLEV, LSC-TRX, SUS_ETMX_OPLEV

 Touching the PSL pointing IOO-QPD_ANG & POS was a mistake. We lost the reference of the well refined MC input.

 

One and 20 days TRENDS  plot showing the PSL output drift in pitch can be power drop

However initial pointing is amazingly good. ( I wonder about the lens in front of the qpd ?)

  9353   Wed Nov 6 14:47:41 2013 SteveFrogsLSCDin connectors added at 1Y2

 The north side of the LSC rack is full. I installed more DIN connectors with fuses on the south side of the rack 1Y2

The access to this may be a little bit awkward. You just remove the connector, wire it and put it back in.

 

  9358   Thu Nov 7 08:57:20 2013 SteveUpdateIOOPSL pointing monitoring

 The qpd sees the power drop as position change.

The laser  monitoring screen shows little changes of the Innolight 2W output. See elog 9292 to compare

So why does the PMC downgrade if the laser output is stationary ?

The PMC-T power is down to 0.75V  The auto locker does maximize power output.

It needs a manual alignment touch up.

 

 

  9359   Thu Nov 7 11:19:04 2013 SteveUpdateVACcrane work POSTPONED to Monday

Quote:

 

The smoke alarms were turned off and surrounding areas were covered with plastic.

The folding I-beam was ground down to be in level with the main beam.

Load bearing cable moved into correct position. New folding spring installed.

Crane calibration was done at 500 lbs at the end of the fully extended jib.

Than we realized that the rotating wheel limit switch stopped working.

This means that the crane is still out of order.

 

 New limit switch will be installed tomorrow morning

Konecranes postponed installation Friday morning Nov. 8

Friday 5pm : Konacranes promising to be here 8am Monday, Nov 11

It was rescheduled on Tuesday again for Wednesday, Nov 13

  9372   Wed Nov 13 08:46:03 2013 SteveUpdateVACvertex crane repair completed

nullQuote:

 

The smoke alarms were turned off and surrounding areas were covered with plastic.

The folding I-beam was ground down to be in level with the main beam.

Load bearing cable moved into correct position. New folding spring installed.

Crane calibration was done at 500 lbs at the end of the fully extended jib.

Than we realized that the rotating wheel limit switch stopped working.

This means that the crane is still out of order.

 

New limit switch installed and tested. The crane is back in full operational mode. Two spare limit switches on hand.

  9374   Wed Nov 13 11:30:07 2013 SteveUpdatePEM particle counter MOVED
Particle counter moved from the top of IOOC to the east wall
  9384   Thu Nov 14 11:41:19 2013 SteveUpdateSUSPRM sensors effected by IR

 IR off for 11 minutes. The PRM  face sensors are effected. The PRM side and the rest of the SUS OSEMS are not effected.

 

  9386   Thu Nov 14 14:35:12 2013 SteveUpdateSUSIR effect on MC and PRM sensors

 Sorry to say but MC1, MC2, MC3 and PRM face OSEMS are having the same problem of leaking IR into the sensors

The PMC was not locked for 11 minutes on this plot.

 

  9388   Fri Nov 15 08:01:20 2013 SteveUpdateSUSETMY damping restored

ETMY sus damping restored

  9389   Fri Nov 15 09:24:41 2013 SteveUpdateIOOWFS with beam dumps

This is a proposal to move WFSs such way that their reflected beam can be trapped.

Later ps: Nic will take care of the Gouy phase telescopes.

  9392   Fri Nov 15 10:31:45 2013 SteveUpdateISSSR560 ISS loop connection

Quote:

Quote:

We have implemented an SR560-based ISS loop using the AOM on the PSL table. This is a continuation of the work in 40m:9328.

We dumped the diffracted beam from the AOM onto a stack of razor blades. This beam is not terribly well separated from the main beam, so the razor blades are at a very severe angle. Any alternatives would have involved either moving the AOM or attempting to dump the diffracted beam somewhere on the PMC refl path. We trimmed the RF power potentiometer on the driver so that with 0.5 V dc applied to the AM input, about 10% of the power is diverted from the main beam.

We ran the PMC trans PD into an AC-coupled SR560. To shape the loop, we set SR560 to have a single-pole low- pass at 300 Hz and an overall gain of 5×104. We take the 600 Ω output and send it into a 50 Ω feed-through terminator; this attenuates the voltage by a factor of 10 or so and thereby ensures that the AOM driver is not overdriven.

The AOM driver's AM input accepts 0 to 1 V, so we add an offset to bias the control signal. The output of the 50 Ω feedthrough is sent into the 'A' input of a second SR560 (DC coupled, A − B setting, gain 1, no filtering). Using a DS345 function generator, a 500 mV offset is put into the 'B' input (the function generator reads −0.250 V because it expects 50 Ω input). The 50 Ω output of this SR560 is sent into the AOM driver's AM input.

A measurement of suppressed and unsuppressed RIN is attached. We have achieved a loop with a bandwidth of a few kilohertz and with an in-loop noise suppression factor of 50 from 100 Hz to 1 kHz. This measurement was done using the PMC trans PD, so this spectrum may underestimate the true RIN.

 A small followup measurement. Here are spectra of the MC trans diode with and without the ISS on. The DC value of the diode (in counts) changed from 17264.2 (no ISS) to 17504.3 (with ISS), but I didn't account for this change in the plot.

There is a small inkling of benefit between 100Hz and 1kHz. Above about 100Hz, the RIN is suppressed to about the noise level of this measurement. Below 100Hz there is no change, which probably means that power fluctuations are introduced downstream of the AOM, which argues for an outer-loop ISS down the road.

Atm #2 is in units of RIN.

 I have disconnected the cable from the SR560 to LSC -ch8 for 15minutes this morning. It is moved from the floor to the top of the chambers as preparation for 40m tour. The SR560 seems to be overloading.

The  ISS servo is off according to the MEDM screen. Why MC-T plot showing zero?  The MC was happy yesterday.

 

  9398   Mon Nov 18 16:39:38 2013 SteveUpdateSUSPRM pictures

PRM is aligned. IFO is not locked. It is just flashing, including arms. Olympus SP570UZ camera used without IR blocker. Note: PRM side OSEM does not show IR effect.

I will take more pictures with IOO IR blocked and HeNe oplev blocked  tomorrow morning.

  9408   Tue Nov 19 11:33:39 2013 SteveUpdateSUSPRM- OSEM side ccd camera is in place

Quote:

Can't we somehow hook up this camera to the MUX with the movie mode?
I think both the MUX and the sensoray are compatible with the color video signal.
Only the old CRT is B/W.

 Watek 902H ccd with Tamron M118FM50 lens is hooked up to MUX  Please be careful! In this set up the lens is close to the view port glass window! 

  9421   Thu Nov 21 16:32:20 2013 SteveUpdateIOOPMC needs a touch of love

 

 The PMC power degrading on this 3 days plot. MC2 -T = 14,200 counts. C1:IOO-MC_TRANS_SUM can not be ploted in dataviewer. The MEDM screen has a valid number.

Initial pointing is not so bad (what does "not so bad" mean ???)

C1iscey comes and goes again.

 

  9422   Fri Nov 22 09:54:22 2013 SteveUpdateCDSDAQ?

Jamie, I think the computers know that you are away. c1lsc keeps going down.

The short time plots are correct.

  9424   Fri Nov 22 16:32:08 2013 SteveUpdateVACRGA scan at day 108

Quote:

 

 Valve configuration: Vacuum Normal

 

  9432   Mon Dec 2 14:24:10 2013 SteveUpdateCDScomputer problems

Rack 1x6 is very noisy.

 SunFire X4600 computer: FB (directly below Megatron) has it's yellow warning light on. It must be loosing one of it's  fan bearings.

 

Jetstore's error message: IDE channel #2 reading error

  9435   Tue Dec 3 07:42:23 2013 SteveUpdateCDSc1iscex timing problem mysteriously disappears??? (thanksgiving miracle???)

Quote:

Steve was trying to do something to it this morning, but I'm not exactly clear on what it was.  Maybe that helped?  Steve, can you tell us what you were trying to do this morning?

 I was trying to repeat  elog 9007  I did only get to line 2 of the Solution by Koji when Ottavia shut down, where I was working. This was all what I did.

  9443   Thu Dec 5 11:06:43 2013 SteveUpdateCameras viewport video ccd cameras are all covered

Quote:

Quote:

Can't we somehow hook up this camera to the MUX with the movie mode?
I think both the MUX and the sensoray are compatible with the color video signal.
Only the old CRT is B/W.

 Watek 902H ccd with Tamron M118FM50 lens is hooked up to MUX  Please be careful! In this set up the lens is close to the view port glass window! 

 PRM-BS & Faraday video ccd  cameras  covered as shown. These thin wall metalized ducts are perfect for blocking light in both direction.

They are very light and give you easy access to adjustment.

  9445   Thu Dec 5 16:20:26 2013 SteveUpdateCDSglitches are gone

Quote:

c1scy had frequent time-over. This caused the glitches of the OSEM damping servos.

Today Eric Q was annoyed by the glitches while he worked on the green PDH inspection at the Y-end.

In order to mitigate this issue, low priority RFM channels are moved from c1scy to c1tst.
The moved channels (see Attachment 1) are supposed to be less susceptible to the additional delay.

This modification required the following models to be modified, recompiled, reinstalled, and restarted
in the listed order:
c1als, c1sus, c1rfn, c1tst, c1scy

Now the models are are running. CDS status is all green.
The time consumption of c1scy is now ~30us (porevious ~60us)
(see Attachment 2)

I am looking at the cavity lock of TEM00 and I have witnessed no glitch any more.
In fact, the OSEM signals have no glitch. (see Attachment 3)

We still have c1mcs having regularly time-over. Can I remove the WFS->OAF connections temporarily?

 Koji cleaned up very nicely.

  9446   Fri Dec 6 10:03:07 2013 SteveUpdateSUSIR effect on MC sensors only

Quote:

 Sorry to say but MC1, MC2, MC3 and PRM face OSEMS are having the same problem of leaking IR into the sensors

The PMC was not locked for 11 minutes on this plot.

 

 The PRM sensors are no longer effected by IR. What changed? The MC still does.

  9448   Fri Dec 6 15:57:41 2013 SteveUpdateIOObeam dumps for PSL pointing monitoring

Quote:

Since the pointing has gone bad again, I went to the PSL to investigate. Found some bad things and removed them:

1) There was a stopped down iris AGAIN in the main beam path, after the newly installed mirror mount. I opened it. Stop closing irises in the beam path.

2) The beam dump for the IOO QPD reflection was just some black aluminum. That is not a real dump. I removed it. We need two razor blade dumps for this.

3) There was an ND filter wheel (???) after one of the PMC steering mirrors. This is not good noise / optics practice. I removed it and dumped the beam in a real dump. No elog about this ?!#?

 

The attached trend shows the last 20 days. The big step ~2 weeks ago is when Steve replaced the steering mirror mount with the steel one. I don't understand the drift that comes after that.

 

Today I also spent ~1 hour repairing the Aldabella laptop. Whoever moved it from the PSL area to the SP table seems to have corrupted the disk by improper shutdown. Please stop shutting the lid and disconnecting it from the AC power unless you want to be fixing it. Its now running in some recovery mode. Lets leave it where it is next to the PSL and MC1.

I steered the MC suspensions back to where they were on the trends before the PSL mirror mount swap and then aligned the PSL beam into it by touching the last 2 steel mounts. Once the alignment was good without WFS, I centered the beams on the IOO QPDs. If it behaves good overnight, I will center the unlocked beams on the MC WFS.

 

Please stay off the PSL for a couple days if you can so that we can watch the drift. This means no opening the doors, turning on the lights, or heavy work around there.

 IOO pointing monitoring qpds received razor beam dumps on their refs.

The Pos QPD was rotated and recentered.

The Ang QPD was left untouched.

TREND plot of 23 days is attached.

  9451   Mon Dec 9 11:44:08 2013 SteveUpdateendtable upgradeETMY optable got new beam dump

Aluminum shield replaced by razor beam dump.

  9452   Tue Dec 10 10:07:01 2013 SteveUpdateIOOmore beam traps

 New razor beam dump installed to trap reflected beam of the input vacuum window.

 

  9458   Thu Dec 12 17:22:07 2013 SteveUpdateGeneralrack power supplies checked

Instrument rack power supplies checked and labeled at present loads.

The vacuum rack Sorensen is running HOT! Their is only 0.3A load at 24V There is plenty of space around it.

It is alarming to me because all vacuum valve positions are controlled by this 24V

  9467   Fri Dec 13 16:34:20 2013 SteveUpdateVACPS will be replaced next week

Quote:

Instrument rack power supplies checked and labeled at present loads.

The vacuum rack Sorensen is running HOT! Their is only 0.3A load at 24V There is plenty of space around it.

It is alarming to me because all vacuum valve positions are controlled by this 24V

 The temperature went down to room temp with temporary fan in the back. Voltage and current are stable.

Regardless, it will be replaced early next week.

ELOG V3.1.3-