  16994   Tue Jul 12 19:46:54 2022 PacoSummaryALSHow (not) to take NPRO PZT transfer function

[Paco, Deeksha, rana]

Quick elog for this evening:

  • Rana disabled MC servo .
  • Slow loop also got disengaged.
  • AUX PSL beatnote is best taken with *free running lasers* since their relative frequency fluctuations are lowest than when locked to cavities.
  • DFD may be better to get PZT transfer funcs, or get higher bandwidth phase meter.
  • Multi instrument to be done with updated moku
  • Deeksha will take care of updated moku
  16997   Wed Jul 13 12:49:25 2022 PacoSummarySUSSUS frozen

[Paco, JC, Yuta]

This morning, while investigating the source of a burning smell, we turned off the c1SUS 1X4 power strip powering the sorensens. After this, we noticed the MC1 refl was not on the camera, and in general other vertex SUS were misaligned even though JC had aligned the IFO in the morning to almost optimum arm cavity flashing. After a c1susaux modbusIOC service restart and burt restore, the problem persisted.

We started to debug the sus rack chain for PRM since the oplev beam was still near its alignment so we could use it as a sensor. The first weird thing we noticed was that no matter how much we "kicked" PRM, we wouldn't see any motion on the oplev. We repeatedly kicked UL coil and looked at the coil driver inputs and outputs, and also verified the eurocard had DC power on which it did. Somehow disconnecting the acromag inputs didn't affect the medm screen values, so that made us suspicious that something was weird with these ADCs.


Because all the slow channels were in a frozen state, we tried restarting c1susaux and the acromag chassis and this fixed the issue.

  17007   Fri Jul 15 19:13:22 2022 PacoSummaryLSCFPMI with REFL/AS55 demod phase adjust

[Yuta, Paco]

  • We first zero the offsets in ASDC, AS55, REFL55, POX11, and POY11 when PSL shutter is closed.
    • After this, we checked the offsets with only ITMX aligned. Some of RFPDs had ~2 counts of offsets, which indicate some RFAM of sidebands, but we decided not to tune Marconi frequencies since the offsets were small enough.
  • We went over the demod phases for AS55, REFL55, POX11, and POY11.
    • For POX11/POY11 first we just minimized the Q in each locked XARM/YARM individually. The newfound values were
      • C1:LSC-POX11_PHASE_R = 106.991
      • C1:LSC-POY11_PHASE_R = -12.820
    • Then we misaligned the XARM by getting rid of the MICH fringe in the ASDC port with ITMX yaw offset, and locked YARM using AS55_Q and REFL55_I and found the demod phase that minimized the AS55_I and REFL55_Q. The newfound values were
      • C1:LSC-AS55_PHASE_R = -65.9586
      • C1:LSC-REFL55_PHASE_R = -78.6254
    • Repeating the above, but now misaligning YARM with ITMY yaw offset, locking XARM with AS55_Q and REFL55_I, we found the demod phases that minimized AS55_1 and REFL55_Q. The newfound values were
      • C1:LSC-AS55_PHASE_R = -61.4361
      • C1:LSC-REFL55_PHASE_R = -71.0434
  • The above demod phases difference, Schnupp asymmetry between X and Y were measured. We repeated the measurement three times to derive the error.
    • Optimal demod phase difference between X arm and Y arm for both AS55 and REFL55 were measured to be -4.5 +/- 0.1 deg, which means that lx-ly = 3.39 +/- 0.05 cm (Marconi frequency: 11.066195 MHz).
  • We measured the gain difference between AS55_Q and POX11/POY11 = -0.5
  • We measured the gain difference between REFL55_I and POX11/POY11 = -2.5

After this, we locked DARM, CARM and MICH using POX11_I, POY11_I and AS55 error signals respectively, and actuating on ETMX, MC2, and BS with NO TRIGGERS (but FM triggers were on for boosts as usual). Under this condition, FM5 is used for lock acquisition, and FM1, FM2, FM3, FM6 are turned on with FM triggers. No FM4 was on. We also noticed:

  • CARM FM6 "BounceRoll" is slightly different than "YARM" FM6 "Bounce". The absent roll resonant gain actually makes it easier to control the CARM, we just had to use YARM filter for locking it.
  • When CARM is controlled, we often just kick the ETMX to bring it near resonance, since the frequency noise drops and we otherwise have to wait long.
  17012   Mon Jul 18 16:39:07 2022 PacoSummaryLSCFPMI locking procedure using REFL55 and AS55

[Yuta, Paco]

In summary, we locked FPMI using REFL55_I, REFL55_Q, and AS55_Q. The key to success was to mix POX11_I and POY11_I in the right way to emulate CARM/DARM, and to find out the correct demodulation phase for AS55.


Procedure

  1. Close PSL shutter and zero offsets in AS55, REFL55, POX11, POY11, and ASDC
    • For ASDC run python3 resetOffsets.py -c C1:LSC-ASDC_IN1, otherwise use the zer offsets on I and Q inputs from the RFPD medm screen.
  2. Lock XARM/YARM using POX/POY to tune demodulation phase.
    • Today, the demode phase in POX11 changed to 104.801, and POY11 to -11.256 deg.
  3. XARM and YARM are used in the following configuration
    • INMAT
      • 0.5 * POX11_I - 0.5 * POY --> XARM
      • 0.5 * POX + 0.5*POY --> YARM
      • REFL55_Q --> MICH (** this should be turned on after POX11/POY11)
    • LSC Filter gains
      • XARM = 0.012
      • YARM = 0.012
      • MICH = +40 (note the sign flip from last time)
    • OUTMAT
      • XARM --> 0.5 * ETMX - 0.5 * ETMY
      • YARM --> MC2
      • MICH --> BS
    • UGFs (sanity check)
      • XARM (DARM) ~ 100 Hz
      • YARM (CARM) ~ 200 Hz
      • MICH (MICH) ~ 40 Hz
  4. Run MICHOpticalGainCalibration.ipynb to see if ASDC vs REFL55_Q looks nice (ellipse in the XY plot), and find any residual offset in REFL55_Q.
    • If the plot doesn't look nice in this regard, the IFO needs to be aligned.
  5. Sensing matrix for CARM/DARM and MICH.
    • With the DARM, CARM and MICH lines on, verify the demod error signals look ok both in mag and phase.
    • For example, we found that CARM error signals were correctly represented by either 0.5 * POX11_I + 0.5 * POY11_I or 0.5 * REFL55_I.
    • Similarly, we found that DARM error signal was correctly represented by either 0.5 * POX11_I - 0.5 * POY11_I or 2.5 * AS55_Q.
    • To find this, we minimized CARM content in AS55_Q, as well as CARM content in REFL55_Q.
  6. We acquired the lock by re-configuring the error point as below:
    • INMAT
      • 0.5*REFL55_I --> YARM (CARM)
      • 2.5 * AS55_Q --> XARM (DARM)
    • During the hand-off trials, we repeatedly ran the sensing matrix and UGF measurements while stopping at various intermediate mixed error points to check how the error signal calibrations changed if at all.
      • Attachment #1 shows the DARM OLTF using POX/POY (blue), only with CARM handoff (green), and after DARM handoff (red)
      • Attachment #2 shows the CARM OLTF using POX/POY (blue), only with CARM handoff (green), and after DARM handoff (red)
      • Attachment #3 shows the MICH OLTF using POX/POY (blue), only with CARM handoff (green), and after DARM handoff (red)
    • The sensing matrix after handoff is below:
Sensing Matrix with the following demodulation phases
{'AS55': 192.8, 'REFL55': 95.63177865911078, 'POX11': 104.80089727128349, 'POY11': -11.256509422276006}
Sensors          	           DARM     	           CARM     	            MICH     	
C1:LSC-AS55_I_ERR_DQ	5.09e-02 (89.6761 deg)	2.03e-01 (-114.513 deg)	1.28e-04 (-28.9254 deg)	
C1:LSC-AS55_Q_ERR_DQ	4.78e-02 (88.7876 deg)	3.61e-03 (-68.7198 deg)	8.34e-05 (-39.193 deg)	
C1:LSC-REFL55_I_ERR_DQ	5.18e-02 (-92.2555 deg)	1.20e+00 (65.2507 deg)	1.15e-04 (-102.027 deg)	
C1:LSC-REFL55_Q_ERR_DQ	1.81e-04 (59.0854 deg)	1.09e-02 (-114.716 deg)	1.77e-05 (-23.6485 deg)	
C1:LSC-POX11_I_ERR_DQ	8.51e-02 (91.2844 deg)	4.77e-01 (67.1709 deg)	7.97e-05 (-72.5252 deg)	
C1:LSC-POX11_Q_ERR_DQ	2.63e-04 (114.584 deg)	1.32e-03 (-113.505 deg)	2.10e-06 (118.146 deg)	
C1:LSC-POY11_I_ERR_DQ	1.58e-01 (-88.9295 deg)	6.16e-01 (67.6098 deg)	8.71e-05 (172.73 deg)	
C1:LSC-POY11_Q_ERR_DQ	2.89e-04 (-89.1114 deg)	1.09e-03 (70.2784 deg)	3.77e-07 (110.206 deg)	

Lock gpstimes:

  1. [1342220242, 1342220260]
  2. [1342220420, 1342220890]
  3. [1342221426, 1342221574]
  4. [1342222753, 1342223230]

Sensitivity estimate (NANB)

Using diaggui, we look at the AS55_Q error point and the DARM control point (C1:LSC-XARM_OUT). We roughly calibrate the error point using the sensing matrix element and actuation gain at the DARM oscillator freq 4.78e-2 / (10.91e-9 / 307.880^2). The control point is calibrated with a 0.95 Hz SUS pole. Attachment #4 shows the sensitivity estimate.

  17021   Wed Jul 20 11:58:45 2022 PacoSummaryGeneralJenne laser kaput?

[Paco, Yehonathan, JC]

We were trying to setup the Jenne laser to characterize the response of three 1811s that Yehonathan is using for his WOPA experiment (in QIL). We hooked up a ~ 5 VDC power supply to the bias tee and looked to see if there was any DC response in the REF PD. We used a DB9 breakout board and a DB9 cable, and saw some current being drawn. The DC current was a bit too high (500 mA), so we turned the DC voltage off, and realized the VDC power was reversed, probably along the DB9 cable which we didn't check before. As we flipped the power supply leads and turned power back on, we could no longer see any current even though the voltage was now right (or was it???). We would like to debug this laser, and continue using it if it still works (!), but there is negligible documentation either here or in the wiki, so if there are any known places to look at it would be helpful to know them.

  17022   Wed Jul 20 14:12:07 2022 PacoSummaryGeneralJenne laser kaput!

[Koji, Yehonathan, Paco]

Koji pointed out that this laser was always driven with a current driver (which was not nearby), and after finding it on one of the rolling carts, we hooked up the system but found that the laser driver displayed open circuit near the usual 20mA operating point. We therefore have to conclude that this laser is no more. We will look for a reasonable replacement.

Quote:

[Paco, Yehonathan, JC]

We were trying to setup the Jenne laser to characterize the response of three 1811s that Yehonathan is using for his WOPA experiment (in QIL). We hooked up a ~ 5 VDC power supply to the bias tee and looked to see if there was any DC response in the REF PD. We used a DB9 breakout board and a DB9 cable, and saw some current being drawn. The DC current was a bit too high (500 mA), so we turned the DC voltage off, and realized the VDC power was reversed, probably along the DB9 cable which we didn't check before. As we flipped the power supply leads and turned power back on, we could no longer see any current even though the voltage was now right (or was it???). We would like to debug this laser, and continue using it if it still works (!), but there is negligible documentation either here or in the wiki, so if there are any known places to look at it would be helpful to know them.

 

  17024   Wed Jul 20 18:07:52 2022 PacoUpdateBHDBHD MICH test

[Paco, Yuta, JC]

We did some easy tests on the BHD readout in preparation for BHD MICH. With the arm cavities and LO beam misaligned, but the MICH aligned, we measured the transfer function from C1:LSC-DCPD_A_OUT to C1:LSC-DCPD_B_OUT to get a rough estimate of the gain balance: 1.8 * DCPD_A = DCPD_B. We then locked MICH using REFL55_Q and looked at

  • A=C1:LSC-DCPD_A_OUT
  • B=C1:LSC-DCPD_B_OUT
  • 1.8 * A - B (which we encoded using C1:LSC-PRCL_A_IN1)
  • 1.8 * A + B (which we encoded using C1:LSC-PRCL_B_IN1)

namely the DCPD BHD signals. After turning the MICH_OSC on (2000 gain @ 311.1 Hz), we took some power spectra under the following three configurations:

  1. LO misaligned, no MICH offset.
  2. LO overlap, no MICH offset.
  3. LO overlap and MICH offset.

For 1. the expectation was that since LO is misaligned and the AS port is dark, we would get no signal. In 2., however both A and B would might see some incoherent signal, but still no MICH. Finally in 3. all signals should be able to see MICH, including A-B. Attachment #1 shows the measurements 1, 2, and 3 (offset = -5.0). Then, with increasing offset values, the BHD MICH signals increased as well; discussion to follow.

  17030   Mon Jul 25 09:05:50 2022 PacoSummaryGeneralTesting 950nm laser found in trash pile

[Paco, Yehonathan]

==== Late elog from Friday ====

Koji provided us with a QFLD-950-3S (QPHOTONICS) salvaged from Aidan's junk pile (LD is alive according to him). We tested the Jenne laser setup with this just to decide if we should order another one, and it worked.

The laser driver anode and cathode pins (8/9, 4/5 respectively) on the rear DB9 port from the  ILX Lightwave LDX-3412 driver were connected to the corresponding anode and cathode pins in the laser package (5, and 9; note the numbers are reversed between driver and laser). Then, interlock pins 1 and 2 in the driver were shorted to enable operation. This is all illustrated in Attachments #1-2.

After setting a limit of 27.6 mA current in the driver, we slowly increased the actual current to ~ 19 mA until we could see light on a beam card. We can go ahead and get a 1060 nm replacement.

  17037   Tue Jul 26 20:54:08 2022 PacoUpdateBHDBHD MICH test - LO phase control

[Yuta, Paco]


TL;DR Successfully controlled LO phase, and did BHD-MICH readout with various MICH offsets and LO phases.


Today we implemented a DCPD based LO phase control. First, we remeasured the balancing gain at 311.1 Hz (the MICH oscillator freq) and combined C1:HPC-DCPD_A_OUT with C1:HPC-DCPD_B_OUT to produce the balanced homodyne error signal (A-B). We feed this error signal to C1:HPC-LO_PHASE_IN1 and for the main loop filters we simply recycled the LSC-MICH loop filters FM2 through FM5 (we also copied FM8, but didn't end up using it much). Then, we verified the LO phase can be controlled by actuating either on LO1 or LO2. For LO2, we added an oscillator in the HPC LOCKINS at 318.75 Hz (we kept this on at 1000 counts for the measurements below).

The LO phase control was achieved with a loop gain in the range of 10-30 (we used 20), no offset, and FM4, and FM5 engaged. FM2 can be added to boost, but we usually skipped FM3. Then, we went through a set of measurements similar to the ones described in a previous elog. A key difference with respect to the measurements from before is that we locked MICH using AS55Q (as opposed to REFL55Q). This allowed us to reach higher MICH offsets without losing lock. After turning on the MICH oscillator at 3000 counts, we looked at:

  1. LO misaligned + MICH at dark fringe (offset = -21).
    • Here, we don't expect to see any MICH signal and indeed we don't, except for a small residual peak from perhaps a MICH offset or slightly imbalanced PDs.
  2. LO aligned, but uncontrolled + MICH at dark fringe (offset = -21).
    • Here we would naively expect MICH to show up in A-B, but because of the uncontrolled LO phase, we mostly see the noise baseline (mostly from LO RIN? ...see measurement 3) under which this signal is probably buried. Indeed, the LO fringe increased noise in A, B, and A-B but not in A+B. This is nice. yes
  3. LO aligned, but uncontrolled + MICH with dc readout (offset = +50).
    • Here we expected the MICH signal to show up due to the large offset, and we can indeed see it in A, B, and A+B, but not in A-B. Nevertheless we see almost exactly the same noise level even though we allow some AS light into the BHD readout, so maybe the noise observed in the A-B channel from measurements 2 and 3 is mostly from LO RIN. This needs further investigation...
  4. LO aligned, controlled at no offset + MICH with dc readout (offset = +50).
    • In general here we expected to see a noise reduction in the A-B channel since the LO fringe is stable, and a MICH signal should appear. Furthermore, since LO phase is under control, we expect the LO2 Oscillator to appear which it does for this and the following measurements. Because of the relative freedom, we tried this measurement in two cases:
      1. When feeding back to LO1
        • We actually see MICH in the A-B channel, as expected, after the noise level dropped by ~ 5. We also observed small sidebands +- 1 Hz away from the MICH peak, probably due to local damping in either LO or AS paths.
      2. When feeding back to LO2
        • We also see MICH here, with a slightly better drop in noise (relative to feeding back to LO1). Sidebands persisted here, but around at +- 2 Hz.
  5. LO aligned, controlled (offset = 10) + MICH with dc readout (offset = +50). *
    • Here, we expected the A-B MICH content to increase dramatically, and indeed it does after a little tuning of the LO phase heart. The noise level decreased slightly because LO phase noise is decreased around the optimal point.
  6. LO aligned, controlled (offset = 20) + MICH with dc readout (offset = +30). *
    • Here, we naively expected A+B MICH content to decrease, but A-B remain constant. In order to see this we tried to keep the balance between the offsets, but this was hard. We don't really see much of this effect, so this also needs further investigation. As long as we keep controlling the LO phase using the DCPDs because the offsets tend to reduce the error signal we will have a harder time.

* For these measurements we actuated on LO2 to keep the LO phase under control.

Note that the color code above corresponds to the traces shown in Attachment #1.


What's next?

  • Alignment of LO and AS might be far from optimized, so it should be tried more seriously.
  • What's the actual LO power? How does it compare with AS power at whatever MICH offsets?
  • Try audio dither LO phase control.
    • With MICH offset.
    • Without MICH offset, double demod (after dolphin fix crying)
  17102   Wed Aug 24 12:02:24 2022 PacoUpdateSUSITMX SUS is sus UL glitches?

[Yehonathan, Paco]

This morning, while attempting to align the IFO to continue with noise-budgeting, we noted the XARM lock was not stable and showed glitches in the C1:LSC-TRX_OUT (arm cavity transmission). Inspecting the SUS screens, we found the ULSEN rms ~ 6 times higher than the other coils so we opened an ndscope with the four face OSEM signals and overlay the XARM transmission. We immediately noticed the ULSEN input is noisy, jumping around randomly and where bigger glitches correlated with the arm cavity transmission glitches. This is appreciated in Attachment #1.


Signal chain investigation

We'll do a full signal investigation on ITMX SUS electronics to try and narrow down the issue, but it seems the glitches come and go... Is this from the gold satamp box? ...

  17104   Thu Aug 25 15:24:06 2022 PacoHowToElectronicsRFSoC 2x2 board -- fandango

[Paco, Chris Stoughton, Leo -- remote]

This morning Chris came over to the 40m lab to help us get the RFSoC board going. After checking out our setup, we decided to do a very basic series of checks to see if we can at least get the ADCs to run coherently (independent of the DACs). For this I borrowed the Marconi 2023B from inside the lab and set its output to 1.137 GHz, 0 dBm. Then, I plugged it into the ADC1 and just ran the usual spectrum analyzer notebook on the rfsoc jupyter lab server. Attachment #1 - 2 shows the screen captured PSDs for ADCs 0 and 1 respectively with the 1137 MHz peaks alright.

The fast ADCs are indeed reading our input signals.


Before this simple test, we actually reached out to Leo over at Fermilab for some remote assistance on building up our minimally working firmware. For this, Chris started a new vivado project on his laptop, and realized the rfsoc 2x2 board files are not included in it by default. In order to add them, we had to go into Tools, Settings and add the 2020.1 Vivado Xilinx shop board repository path to the rfsoc2x2 v1.1 files. After a little bit of struggling, uninstalling, reinstalling them, and restarting Vivado, we managed to get into the actual overlay design. In there, with Leo's assistance, we dropped the Zynq MPSoC core (this includes the main interface drivers for the rfsoc 2x2 board). We then dropped an rf converter IP block, which we customized to use the right PLL settings. The settings, from the System Clocking tab were changed to have a 409.6 MHz Reference Clock (default was 122.88 MHz). This was not straightforward, as the default sampling rate of 2.00 GSPS was not integer-related so we had to also update that to 4.096 GSPS. Then, we saw that the max available Clock Out option was 256 MHz (we need to be >= 409.6 MHz), so Leo suggested we dropped a Clocking Wizard block to provide a 512 MHz clock input for the rfdc. The final settings are captured in Attachment # 3. The Clocking Wizard was added, and configured on its Output Clocks tab to provide a Requested Output Freq of 512 MHz. The finall settings of the Clocking wizard are captured in Attachment #4. Finally, we connected the blocks as shown in Attachment #5.

We will continue with this design tomorrow.

  17133   Tue Sep 6 17:39:40 2022 PacoUpdateSUSLO1 LO2 AS1 AS4 damping loop step responses

I tuned the local damping gains for LO1, LO2, AS1, and AS4 by looking at step responses in the DOF basis (i.e. POS, PIT, YAW, and SIDE). The procedure was:

  1. Grab an ndscope with the error point signals in the DOF basis, e.g. C1:SUS-LO1_SUSPOS_IN1_DQ
  2. Apply an offset to the relevant DOF using the alignment slider offset (or coil offset for the SIDE DOF) while being careful not to trip the watchdog. The nominal offsets found for this tuning are summarized below:
Alignment/Coil Step sizes
  POS PIT YAW SIDE
LO1 800 300 300 10000
LO2 800 300 400 -10000
AS1 800 500 500 20000
AS4 800 400 400 -10000
  1. Tune the damping gains until the DOF shows a residual Q with ~ 5 or more oscillations.
  2. The new damping gains are below for all optics and their DOFs, and Attachments #1-4 summarize the tuned step responses as well as the other DOFs (cross-coupled).
Local damping gains
  POS PIT YAW SIDE
LO1 10.000 5.000 3.000 40.000
LO2 10.000 3.000 3.000 50.000
AS1 14.000 2.500 3.000 85.000
AS4 15.000 3.100 3.000 41.000

Note that during this test, FM5 has been populated for all these optics with a BounceRoll (notches at 16.6, 23.7 Hz) filter, apart from the Cheby (HF rolloff) and the 0.0:30 filters.

  17142   Thu Sep 15 21:12:53 2022 PacoUpdateBHDLO phase "dc" control

Locked the LO phase with a MICH offset=+91. The LO is midfringe (locked using the A-B zero crossing), so it's far from being "useful" for any readout but we can at least look at residual noise spectra.

I spent some time playing with the loop gains, filters, and overall lock acquisition, and established a quick TF template at Git/40m/measurements/BHD/HPC_LO_PHASE_TF.xml

So far, it seems that actuating on the LO phase through LO2 POS requires 1.9 times more strength (with the same "A-B" dc sensing). After closing the loop by FM4, and FM5, actuating on LO2 with a filter gain of 0.4 closes the loop robustly. Then, FM3 and FM6 can be enabled and the gain stepped up to 0.5 without problem. The measured UGF (Attachment #1) here was ~ 20 Hz. It can be increased to 55 Hz but then it quickly becomes unstable. I added FM1 (boost) to the HPC_LO_PHASE bank but didn't get to try it.

The noise spectra (Attachment #2) is still uncalibrated... but has been saved under Git/40m/measurements/BHD/HPC_residual_noise_spectra.xml

  17143   Mon Sep 19 17:02:57 2022 PacoSummaryGeneralPower Outage 220916 -- restored all

Restore lab

[Paco, Tega, JC, Yehonathan]

We followed the instructions here. There were no major issues, apart from the fb1 ntp server sync taking long time after rebooting once.


ETMY damping

[Yehonathan, Paco]

We noticed that ETMY had to much RMS motion when the OpLevs were off. We played with it a bit and noticed two things: Cheby4 filter was on for SUS_POS and the limiter on ULCOIL was on at 0 limit. We turned both off.

We did some damping test and observed that the PIT and YAW motion were overdamped. We tune the gain of the filters in the following way:

SUSSIDE_GAIN 1250->50

SUSPOS_GAIN 200->150

SUSYAW_GAIN 60->30

These action seem to make things better.

  17145   Tue Sep 20 07:03:04 2022 PacoSummaryGeneralPower Outage 220916 -- restored all

[JC, Tega, Paco ]

I would like to mention that during the Vacuum startup, after the AUX pump was turned on, Tega and I were walking away while the pressure decreases. While we were, valves opened on their own. Nobody was near the VAC Desktop during this. I asked Koji if this may be an automatic startup, but he said the valves shouldn't open unless they are explicitely told to do so. Has anyone encountered this before?

Quote:

Restore lab

[Paco, Tega, JC, Yehonathan]

We followed the instructions here. There were no major issues, apart from the fb1 ntp server sync taking long time after rebooting once.


ETMY damping

[Yehonathan, Paco]

We noticed that ETMY had to much RMS motion when the OpLevs were off. We played with it a bit and noticed two things: Cheby4 filter was on for SUS_POS and the limiter on ULCOIL was on at 0 limit. We turned both off.

We did some damping test and observed that the PIT and YAW motion were overdamped. We tune the gain of the filters in the following way:

SUSSIDE_GAIN 1250->50

SUSPOS_GAIN 200->150

SUSYAW_GAIN 60->30

These action seem to make things better.

 

  17150   Wed Sep 21 17:01:59 2022 PacoUpdateBHDBH55 RFPD installed - part I

[Radhika, Paco]

Optical path setup

We realized the DCPD - B beam path was already using a 95:5 beamsplitter to steer the beam, so we are repurposing the 5% pickoff for a 55 MHz RFPD. For the RFPD we are using a gold RFPD labeled "POP55 (POY55)" which was on the large optical table near the vertex. We have decided to test this in-situ because the PD test setup is currently offline.

Radhika used a Y1-1025-45S mirror to steer the B-beam path into the RFPD, but a lens should be added next in the path to focus the beam spot into the PD sensitive area. The current path is illustrated by Attachment #1.

We removed some unused OPLEV optics to make room for the RFPD box, and these were moved to the optics cabinet along Y-arm [Attachment #2].

 


[Anchal, Yehonathan]

PD interfacing and connections

In parallel to setting up the optical path configuration in the ITMY table, we repurposed a DB15 cable from a PD interface board in the LSC rack to the RFPD in question. Then, an SMA cable was routed from the RFPD RF output to an "UNUSED" I&Q demod board on the LSC rack. Lucky us, we also found a terminated REFL55 LO port, so we can draw our demod LO from there. There are a couple (14,15,20,21) ADC free inputs after the WF2 and WF3 whitening filter interfaces.


Next steps

  • Finish alignment of BH55 beam to RFPD
  • Test RF output of RFPD once powered
  • Modify LSC model, rebuild and restart
  17159   Mon Sep 26 11:39:37 2022 PacoUpdateBHDBH55 RFPD installed - part II

[Paco, Anchal]

We followed rana's suggestion for stress relief on the SMA joint in the BH55 RFPD that Radhika resoldered. We used a single core, pigtailed wire segment after cleaning up the solder joint on J7 (RF Out) and also soldered the SMA shield to the RF cage (see Attachment #1). This had a really good effect on the rigidity of the connection, so we moved back to the ITMY table.

We measured the TEST in to RF Out transfer function using the Agilent network analyzer, just to see the qualitative features (resonant gain at around 55 MHz and second harmonic suppression at around 110 MHz) shown in Attachment #2. We used 10kOhm series resistance in test input path to calibrate the measured transimpedance in V/A. The RFPD has been installed in the ITMY table and connected to the PD interface box and IQ demod boards in the LSC rack as before.

Measurement files

  17160   Tue Sep 27 10:50:11 2022 PacoUpdateBHDcalibrated LO phase noise

Locked LO phase to ITMX single bounce beam at the AS port, using the DCPD (A-B) error point and actuating on LO1 POS. For this the gain was tuned from 0.6 to 4.0. A rough Michelson fringe calibration gives a counts to meters conversion of ~0.212 nm/count, and the OLTF looks qualitatively like the one in a previous measurement (~ 20 dB at 1 Hz, UGF = 30 Hz). The displacement was then converted to phase using lambda=1e-6; I'm not sure what the requirement is on the LO phase (G1802014 says 1e-4 rad/rtHz at 1 Hz, but our requirement doc says 1 to 20 nrad/rtHz (rms?)... anyways wit this rough calibration we are still off in either case.

The balancing gain is obvious at DC in the individual DCPD spectra, and the common mode rejection in the (A-B) signal is also appreciable. I'll keep working on refining this, and implementing a different control scheme.

  17161   Wed Sep 28 16:37:26 2022 PacoUpdateBHDcalibrated LO phase noise; update

[yuta, paco]

Update; the high frequency ( > 100 Hz) drop is of course not real and comes from a 4th order LP filter in the HPC demod I filter which I haven't accounted for. Furthermore, we have gone through the calibration factors and corrected a factor of 2 in the optical gain. Then, I also added the CLTF to show in loop and out of loop error respectively. The updated plot, though not final, is in Attachment #1.

  17163   Wed Sep 28 21:54:08 2022 PacoUpdateBHDcalibrated LO phase noise; update

Repeated the LO phase noise measurement, this time with the LO - ITMY single bounce, and a couple of fixes Koji hinted at including:

  1. The DEMOD angle was the missing piece! The previous error point showed lower noise than the individual DCPDs because the demodulation angle had not been checked. I corrected it so that the error point in LO_PHASE control was exactly equal to the LO-ITMY single bounce fringe. With this, the gain on the servo had to be adjusted from 4.00 to 0.12, still using FM4, FM5, and this time also FM8 (BLP600).
  2. Turned off 60 Hz harmonics comb notches on DCPDs, they are unecessary.
  3. Acquired noise spectra down to 0.1 Hz, with 0.03 Hz bin width to increase resolution and identify resonant SUS noise near 1 Hz.

This time, after alignment the fringe amplitude was 500 counts. Attachment #1 shows the updated plot with the calibrated noise spectra for the individual DCPD signals A and B as well as their rms values. Attachment #2 shows the error point, in loop and the estimated out of loop spectra with their rms as well. The peak at ~ 240 Hz is quite noticeable in the error point time series, and dominates the high frequency rms noise. The estimated rms out of loop noise is ~ 9.2 rad, down to 100 mHz.

  17167   Fri Sep 30 20:18:55 2022 PacoUpdateBHDLO phase noise with different actuation points

[Paco, Koji]

We took lo phase noise spectra actuating on the for different optics-- LO1, LO2, AS1, and AS4. The servo was not changed during this time with a gain of 0.2, and we also took a noise spectrum without any light on the DCPDs. The plot is shown in Attachment #1, calibrated in rad/rtHz, and shown along with the rms values for the different suspension actuation points. The best one appears to be AS1 from this measurement, and all the optics seem to show the same 270 Hz (actually 268 Hz) resonant peak.


268 Hz noise investigation

Koji suspected the observed noise peak belongs to some servo oscillation, perhaps of mechanical origin so we first monitored the amplitude in an exponentially averaging spectrum. The noise didn't really seem to change too much, so we decided to try adding a bandstop filter around 268 Hz. After the filter was added in FM6, we turned it on and monitored the peak height as it began to fall slowly. We measured the half-decay time to be 264 seconds, which implies an oscillation with Q = 4.53 * f0 * tau ~ 3.2e5. This may or may not be mechanical, further investigation might be needed, but if it is mechanical it might explain why the peak persisted in Attachment #1 even when we change the actuation point; anyways we saw the peak drop ~ 20 dB after more than half an hour... After a while, we noticed the 536 Hz peak, its second harmonic, was persisting, even the third harmonic was visible.

So this may be LO1 violin mode & friends -

We should try and repeat this measurement after the oscillation has stopped, maybe looking at the spectra before we close the LO_PHASE control loop, then closing it carefully with our violin output filter on, and move on to other optics to see if they also show this noise.

  17171   Mon Oct 3 15:19:05 2022 PacoUpdateBHDLO phase noise and control after violin mode filters

[Anchal, Paco]

We started the day by taking a spectrum of C1:HPC-LO_PHASE_IN1, the BHD error point, and confirming the absence of 268 Hz peaks believed to be violin modes on LO1. We then locked the LO phase by actuating on LO2, and AS1. We couldn't get a stable loop with AS4 this morning. In all of these trials, we looked to see if the noise increased at 268 Hz or its harmonics but luckily it didn't. We then decided to add the necessary output filters to avoid exciting these violin modes. The added filters are in the C1:SUS-LO1_LSC bank, slots FM1-3 and comprise bandstop filters at first, second and third harmonics observed previously (268, 536, and 1072 Hz); bode plots for the foton transfer functions are shown in Attachment #1. We made sure we weren't adding too much phase lag near the UGF (~ 1 degree @ 30 Hz).

We repeated the LO phase noise measurement by actuating on LO1, LO2 and AS1, and observe no noise peaks related to 268 Hz this time. The calibrated spectra are in Attachment #2. Now the spectra look very similar to one another, which is nice. The rms is still better when actuating with AS1.


[Paco]

After the above work ended, I tried enabling FM1-3 on the C1:HPC_LO_PHASE control filters. These filters boost the gain to suppress noise at low frequencies. I carefully enabled them when actuating on LO1, and managed to suppress the noise by another factor of 20 below the UGF of ~ 30 Hz. Attachment #3 shows the screenshot of the uncalibrated noise spectra for (1) unsupressed (black, dashed), (2) suppressed with FM4-5 (blue, solid), and (3) boosted FM1-5 suppression (red).


Next steps:

  • Compare LO-ITMY and LO-ITMX single bounce noise spectra and MICH.
  • Compare DC locking scheme versus BH55 once it's working.
  17206   Mon Oct 24 18:01:00 2022 PacoSummaryBHDBHD actuation measurements

[Yuta, Paco]

Today we calibrated the actuation on BHD suspended optics: LO1, LO2, AS1, AS4.
Actuation transfer functions for these optics look good.


ITMY actuation

For a reference we locked LO-ITMY single bounce using the LSC MICH loop. The error point was BH55_Q, the whitening filter gain was 45 dB, IQ demod rotation angle = 151.061 deg, the servo gain was -10, and the actuation point was ITMY. The measured UGF for this loop was ~ 150 Hz when FM2, 3, 4, 5 and 8 were all enabled. Note FM8 is an elliptic low pass (600 Hz cutoff).


LO1, LO2, AS1, AS4 actuation

We then lock the LO phase by feeding back BH55_Q_ERR to the actuation points under test with exactly the same filters but a servo gain of 0.6 but otherwise we are using the same servo filters FM2, 3, 4, 5 and 8 for this controls. The measured UGFs were all near ~ 70 Hz.

Here we had to be careful not to excite mechanical (?) resonances similar to the previously observed "violin" modes in LO1. In particular, we first noticed unsupressed 816 Hz noise in AS1 was being reinjected by the loop sometimes tripping the local damping loops, so we added bandstop filters at the AS1_LSC output filter bank. The resulting loop was then allowed to increase the gain and turn on FM2 and FM3 (boosts). This was also the case in AS4, where 268 Hz and second + third harmonics appeared to be excited by our feedback control. Finally, AS4 also displayed some mechanical excitation at 96.7 Hz, which seemed too low to be a "violin" mode, and its "Q" factor was not as high. We added a bandstop for this as well.

Attachment #1 shows LO_PHASE OLTFs when actuating in the different optics. By taking the actuation ratios (Attachment #2) with respect to our ITMY actuation reference and which had previously been calibrated to be 4.74e-9 / f^2 m / cts, we now have estimated our BHD suspension actuation calibrations to be:

  • LO1 = 3.14e-8 / f^2 m / cts
  • LO2 = 2.52e-8 / f^2 m / cts
  • AS1 = 3.14e-8 / f^2 m / cts
  • AS4 = 2.38e-8 / f^2 m / cts

This magnitudes are consistent with the expected coil driver ranges (about a factor of 10 difference).

  17209   Tue Oct 25 09:57:34 2022 PacoConfigurationPEMAuto Z on trillium interface board

I pressed the Auto-Z(ero) button for ~ 3 seconds at ~9:55 local (pacific) time on the trillium interface on 1X5.

  17211   Tue Oct 25 14:29:56 2022 PacoSummarySEIEarthquake tripped SUS

[Yuta, Paco, JC]

This eq  potentially tripped ETMY, PR2, PR3, AS1, AS4, SR2, LO1, LO2 suspensions during today's WB meeting. We restored them into normal local damping.


We aligned the arm cavities just to verify things were ok and then moved on to BHD comissioning. No problems spotted so far.

  17212   Tue Oct 25 17:27:11 2022 PacoSummaryBHDLO phase control with RF + audio sidebands

[Yuta, Paco]

Today we locked LO phase with BH55 + Audio dithering


Configuration

We worked with MICH locked using AS55_Q with an offset = 50. Our BH55_Q_ERR is the same as in the previous elog (in this thread). We enabled audio dithering of AS1 to produce 280.54 Hz sidebands (exc gain = 15000). We used ELP80 (elliptic, 4th order lowpass with the second resonant notch at 280.54 Hz) at the BH55_Q_AS1_DEMOD_I output. This allowed us to generate an error signal to feedback into AS1 POS. Attachment #1 shows a screen capture of this configuration.


Lock

We close a loop with the above configuration to lock the LO phase using only filters FM5, FM8 and then optionally boost with FM2. The compromise we had to make because of our phase margin was to achieve UGF ~ 20 Hz (in contrast with ~ 70 Hz used in single bounce). Attachment #2 shows the measured OLTFs for LO_PHASE control using this scheme; the red was the final measured loop, while the blue was our initial reference before increasing the servo gain.

 

  17215   Wed Oct 26 10:28:05 2022 PacoUpdateIOOHEPA ON / WFS ON

Turned HEPA ON this morning at 10:28 local (pacific time) or gpstime = 1350802758. WFS ON right after that. IMC was locked and nominally aligned at this time.

  17216   Wed Oct 26 16:04:12 2022 PacoSummaryBHDLO phase control with RF + audio sidebands

[Yuta, Paco]

Today we again locked the LO phase with BH55 + Audio dithering under a zero-offset MICH


Configuration

We worked with MICH locked using AS55_Q with an offset = 0. Our BH55_Q_ERR is the same as in the previous elog (in this thread).We reduced the MICH offset from 50 to 0 slowly and kept an eye on the BH55 error signals. We realized that at zero offset, most of the error signal was in BH55_I_ERR (why?) so we rotated it back to BH55_Q_ERR (146 deg --> 56 deg). We then looked at the audio demod angle, and optimized it to allocated the error signal in the I quadrature (-15 deg --> 40 deg).


Lock

We close a loop with the above configuration to lock the LO phase using only filters FM5, FM8 and then optionally boost with FM2. The measured UGF ~ 20 Hz similar to the configuration with an offset present; and it seems there is some residual noise at ~ 20 Hz (observed in the residual error signal time trace with ndscope).


Next tasks:

  • Noise budget residual error in this configuration
  • Investigate negative count offset in DCPDs
  • Investigate why does the rotation angle change from single bounce to MICH?
  17221   Wed Nov 2 16:34:56 2022 PacoSummaryCalibrationSingle arm calibration run

[Anchal, Paco]

We added a notch filter on ETMY (the actuation point of the YARM control loop) to inject our calibration line at 575.170 Hz. The excitation is injected using the DARM Oscillator, with an exc. gain of ~ 500 (this gets us a decent > 10 SNR line in the ALS Y beat). With the arm cavity locked to the PSL (~150 Hz control bandwidth), and the aux laser locked to the cavity (~10 kHz control bandwidth) the goal of this run is to calibrate our actuator strength and more importantly to budget its uncertainty. For this we have looked at the ALS beat stability using Allan statistics; we noticed the ALS beatnote frequency fluctuations start to become dominated by 1/f (or divergent noise due to systematic drifts in the YARM loop) after 10 seconds (see Attachment #1) (we have managed to see 30 seconds stability with the HEPAs off and without locking to IMC).

Our prediction is that our demodulated calibration lines will display the least residual rms noise when averaging down to around this time. This is the only reason one would use allan statistics; to quantify the separation between statistical and systematic effects in a frequency measurement. To be continued...


gpstimes:

  • 1351464997 to 1351467139
  • 1351467259 to 1351468221
  • 1351468318 to ...
  17253   Thu Nov 10 17:40:31 2022 PacoSummaryCalibrationCalibration Plan

Plan to calibrate single arm actuation strength

  1. Lock single arm cavity (e.g. YARM)
  2. Lock YAUX laser to arm cavity (actuation point is ETMY)
  3. With the notch on the YARM loop filter (actually on ETMY),
  4. Turn on cal line (e.g. DARM osc) to move ITMY; here the frequency is chosen to be away from 600 Hz (line harmonic) and from violin modes for ITMY (642 Hz). The lower value of 575.17 Hz was chosen to avoid demodulating noise peaks at 455 Hz and 700 Hz.
  5. Get raw YALS beatnote (we chose the demod angle of -35 deg to minimize Q).

The analysis is as follows:

  1. Get demodulated IQ timeseries for the duration of the locks before lowpass filter (C1:CAL-SENSMAT_DARM_BEATYF_I_DEMOD_I_IN1); we are also storing the raw beatnote if we want to do software demodulation.
  2. Look at the allan deviation of I and Q to establish the timescale over which our measurement is dominated by statistical uncertainty -- after this time, the uncertainty is expected to be due to systematic error / drift. In this case as shown by Attachment #1 the time is around 60.6 seconds.
  3. At this frequency and with 500 gain the ITMY coils should be actuating 7.32 pm of amplitude displacement.
  4. The minimum allan deviation does indeed predict the statistical uncertainty limited rms if we look at the power spectra of the demodulated cal line over different time periods (Attachment #2), notice I lowpassed the raw timeseries.
  5. I think the next step is to get the nominal calibration value and repeat the measurment for more than a single cal line.
  6. Roughly from the deviation plot, our fractional beatnote deviation is a proxy for the calibration uncertainty. 1.15e-16 of beatnote stability should translate to a fractional displacement stability of ~4.57e-15 at 60 seconds; giving an ultimate statistical calibration uncertainty of 0.06% at this particular frequency when averaging for this long. It might be interesting to see a calibration frequency dependent allan deviation plot.
  17257   Fri Nov 11 14:15:45 2022 PacoSummaryCalibrationSingle arm cal with 5 lines

I turned all the LSC oscillators on and used the digital demod for BEATYF (fine y als beat) to grab the data. For this I added notches onto SCY ETMY LSC filter banks FM6-10 to account for these lines at 30.92, 211.10, 313.31, 315.17 Hz (basically just reusing the osc models) and adjusted the sensing matrix to actuate on ITMY.

I aligned and locked YARM, and then I aligned and locked the YAUX. The lock seems pretty robust with an avg green transmission GTRY ~ 0.185 counts for TEM00.


Trying to see other lines appear on the BEATYF demod channels, but so far no luck. I scaled down the exc gain from 500 counts (snr ~ 20 at 575 Hz) and verified the notches are working. Since I am unsure of the issue here and WFS tests are happening at 4 today, I decided to take some beat data under different conditions -->

HEPA OFF and PSL Shutter Open

gpstime start = 1352244763 gpstime end = 1352245405

HEPA OFF and

PSL Shutter Closed

gpstime start = 1352245574

gpstime end = 1352246216

HEPA ON and

PSL Shutter Closed

gpstime start = 1352246240 gpstime end = 1352246882
  17264   Mon Nov 14 14:52:56 2022 PacoConfigurationSUSBHD SUS Coil output balance

[JC, Paco]

We installed a steering mirror intersecting the BHD beam path and put the AS beam on the ITMY Oplev QPD (see Attachment #1 for a photo of this temporary hack) . This is done to do coil balancing of AS1/AS4, LO1/LO2. QPD sees ~ 10000 counts when the beam is centered.


[Paco, Yuta]

We follow this procedure -- but with different sensors for all BHD suspension coil output balancing.

AS1/AS4

We dither BUTT first, lock the LO-AS fringe (DC lock), and look at the residual LO_PHASE spectrum to minimize POS coupling. We then unlock, misalign LO beam and look at the hijacked Oplev (ITMY) while dithering POS to minimize PIT and YAW couplings.


LO1/LO2

We dither BUTT first, lock thChangeset summarye LO-AS fringe (DC lock), and look at the residual LO_PHASE spectrum to minimize POS coupling. We then unlock, misalign AS beam and look at the hijacked Oplev (ITMY) PIT/YAW residual noise while dithering POS to minimize PIT/YAW coupling.


Changeset summary

The new coil output gains are summarized in the table below:

Optic / Coil UL UR LR LL
AS1 -0.939 1.040 -1.026 0.995
AS4 -0.9785 0.9775 -1.0695 0.9745
LO1 -0.939 1.003 -1.074 0.984
LO2 -1.051 1.342 -0.976 0.631

Finally, I reverted the hacked QPD setup to restore the ITMY OPLEV.

  17266   Tue Nov 15 11:04:01 2022 PacoSummaryCalibrationSingle arm cal with 5 lines

YALS Hardware inspection

The ALS Cal for ITMY actuation was off by ~ 1000, so I decided I don't trust / understand what this beatnote is seeing. Then, I went in the lab to inspect things;

  1. ALS DFD and DEMOD: The demod box was off no perhaps on purpose, more likely by accident... the switch was off in the rear of the 1U box in the LSC rack... Also, the DB9 cable labeled ALS was disconnected. Fixed both these bugs and verified it worked all the way into cds.
  2. YARM Green injection: Did some re-alignment, and noted the mode was hopping a bit too much (once every couple of seconds) so I rotated the half waveplate before the green PZT steering mirrors by  ~ 8 deg and used the latter to get a GTRY transmission of > 0.320 (counts), about 75% more than last time. Finally I made sure the mode is robustly locked for several minutes.

Beatnote calibration

The factor above may be explained by the bogus signals coming into the beat fine phase channels on the ALS model. After locking the YARM with POY11, and locking the YAUX to the YARM cavity, I turned on the LSC oscillators -- all five of them see Attachment #1 for the screenshot -- and looked for the lines in the C1:ALS-BEATY_FINE_PHASE_OUT channel. Here, again the sensing output matrix was set up to actuate on ITMY, while the ETMY (control point of YARM loop) had all the output notches on. Once all lines were visible in the YAUX beatnote, I had to reduce the LSC filter gain from 0.012 to 0.011 to prevent loop oscillations... Then I recorded the gpstimes below with different conditions.

  • HEPA ON (JC Inside lab)
    • gpstime = 1352575798 to 1352576046
  • HEPA OFF (JC Inside lab)
    • gpstime = 1352576125 to 1352576216
  • HEPA OFF (JC in the control room)
    • gpstime = 1352576641 to 1352577591

Analysis

Basically, only the DARM line was recorded (DQ channs) so I modified the c1cal to store the SIG_OUT and DEMOD_I_IN1 channels for both BEATX and BEATY cal signals. This means I need to repeat this measurement. In the meantime I am also going to try and rerun calibrate the BEAT HZ transfer function.

  17268   Tue Nov 15 17:08:59 2022 PacoUpdateBHDRequest for estimates

[Yehonathan, Yuta, Paco]

We would like to estimate:

  • LO phase sensitivty (for RF55 + audio dither scheme), as a function of RF demod angle (both I and Q); not to be confused with audio dither angle.
  • LO phase sensitivity (for all schemes like in Attachment #2 of this previous post) but with some nonzero MICH offset.
  • LO phase sensitivity (for RF55 + audio dither scheme) but with the uBHDBS (44:56) values from this post.
  17269   Tue Nov 15 17:58:00 2022 PacoConfigurationCamerasPOP camera realignment after IFO alignment

[Paco, Yuta]

I swapped the 1 inch BS and lenses along the POP beam to clear the apertures and avoid clipping this beam. The results are illustrated by the attached pictures; this was done right after Yuta had optimized IFO alignment so it's hopefully a good reference from now on. Yuta also tuned the alignment of BHDC path in ITMY table, which mostly improved the alignment to DCPD A (90-ish counts improved to 100-ish counts with ITMY single bounce).

  17278   Thu Nov 17 12:24:48 2022 PacoConfigurationCamerasITMX Camera -- attempt at fix

I found that an old BNC cable for ITMXF video existed so I first tried swapping both ends of the cable, one on the ITMX viewport and the other one in the video MUX input in the rear. This didn't fix the issue.

I searched around in the CCD cabinet by XARM and found an identical analog camera so I swapped it and got the same image ...

I then searched for a AC/DC supply cable, but couldn't find one.

Quote:

Coming in this morning, I found ITMX Camera malfunctioning.

 

  17303   Wed Nov 23 14:59:11 2022 PacoSummaryBHDBHD_DIFF sensitivity to BS dither with MICH Offset

[Yuta, Paco, Anchal]

We measured

(a) BHDC_DIFF sensitivity to BS dither for a set of MICH offsets.


Configurations

  • MICH locked with AS55_Q
    • The MICH offset was varied below
  • LO_PHASE locked with BH55_Q
    • Balanced DCPD_A and DCPD_B by applying a digital gain of 1.00 to DCPD_A
    • Changed the BH55 demod angle to 140.07 deg to minimize BH55_I
  • BS dither at 311.1 Hz
    • Use newly added HPC_BS Lockins to readback the demodulated signals

Results & Discussion

The analysis was done with the '/cvs/cds/rtcds/caltech/c1/Git/40m/scripts/CAL/BHD/BHD_DIFFSensitivity.ipynb' notebook.

Attachment #1 shows the main result showing the sensitivity of various demodulated error signals at 311.1 Hz for a set of 21 MICH offsets. We noted that if we didn't randomize the MICH offset scan, we observed a nonzero "zero crossing" for the offset.
Note that, although LO_PHASE loop was always on to control the LO phase to have zero crossing of BH55_Q, actual LO phase is not constant over the measurement, as MICH offset changes BH55_Q zero crossing.
When MICH offset is zero, LO_PHASE loop will control the LO phase to 0 deg (90 deg away from optimal phase), and BHDC_DIFF will not be sensitive to MICH, but when MICH offset is added, BHDC_DIFF start to have MICH sensitivity (measurement is as expected).
For BHDC_SUM, MICH sensitivity is linear to MICH offset, as it should be the same as ASDC, and does not depend on LO phase (measurement is as expected).
For BH55_Q, MICH sensitivity is maximized at zero MICH offset, but reduces with MICH offset, probably because LO phase is also being changed.


  17318   Mon Nov 28 16:58:20 2022 PacoUpdateCDSpypi package added

[Paco, Tega]

I added the pypi package "restoreEpics" to the donatella clone under test. This is required by some of Anchal's scripts that turn on F2A filters as well as other recovery stages during some measurements.

  17319   Mon Nov 28 18:21:50 2022 PacoSummaryBHDBH44 prep

I checked the LSC rack to evaluate what we might need to generate 44 MHz rf in the hypothetical case we go from BH55 to BH44 (a.k.a. double RF demod scheme). There is an 11 MHz LO port labeled +16 dBm (measured 9 Vpp ~ 23 dBm actually) on the left hand side. Furthermore, there is an unused 55 MHz port labeled "Spare 55 LO". I checked this output to be 1.67 Vpp ~ +8.4 dBm. Anyways the 55 MHz doesn't look very nice; after checking it on the spectrum analyzer it seems like lower frequency peaks are polluting it so it may be worth checking the BH55 LO (labeled REFL 55) signal to see if it's better. Anyways we seem to have the two minimum LOs needed to synthesize 44 MHz in case we move forward with BH44.


[Paco, Yuta]

We confirmed the noisy 55 MHz is shared between AS55, BH55 and any other 55 MHz LOs. Looking more closely at the spectrum we saw the most prominent peaks at 11.06 MHz and 29.5 MHz (IMC and PMC nominal PM freqs). This 55 MHz LO is coming all the way from the RF distribution box near the IOO rack. According to this diagram, this 55 MHz LO should have gone through a bandpass filter; interestingly, checking the RF generation box spare 55 MHz the output is *cleaner* and displays ~ 17 dBm level... ??? Will continue investigating when we actually need this RF.

  17326   Wed Nov 30 18:27:07 2022 PacoUpdateGeneralMaking the Jenne laser great again

[Paco]

I picked up the 950 nm laser that Koji tested before and restored the Jenne laser (may long it live).


I changed a couple of things; first I took out the old laser from the hose and noted the new laser has a shorter fiber patch cable, so I replaced the hose with a flexible fiber jacket I found around the X arm storage. I also added a 14 pin DIP socket so the laser is now always showing its part no. and is easy to install/uninstall. Last but not least I replaced the black insulating material right at the output of the laser package because it was decaying badly and spread a bunch of dusty residue all over the place. I used some foam instead.  See various attachments for visual support.

I was careful, testing the connections with a cheap LED to validate the setup before I connected the laser-- this made everything smooth and I finished the work by verifying the laser is outputting light at the other end of the fiber 1x2 coupler. The PD testing with Jenne's resucitated laser should proceed normally now.

RXA: Huzzah!!

  17346   Thu Dec 8 16:21:40 2022 PacoSummaryCalibrationITMY actuation strength cal with 5 lines

[Anchal, Paco]

After debugging the hardware, on gpstime 1354422834 we turned on 5 cal lines on ITMY to test the ALS calibration for the single arm along with our error estimates.

Note: the YARM IR lock lasted > 8 hours, but the GRY transmission dropped twice during the evening and hopped back up, so the phase tracker jumped a couple of times.


Configuration

YAUX laser was locked to the YARM through the analog PDH servo (UGF ~ 2 kHz), YARM was locked to the PSL with POY11 (UGF ~ 200 Hz), and the ALS phase tracker was set to output the beat frequency noise in Hz. HEPA was left on during this measurement. The oscillators were similar to previous instances: gains of 70@211.1Hz, 100@313.31Hz, 100@315.17Hz, 300@575.17Hz and 15@30.92Hz with appropriate notches on ETMY to avoid POY11 loop supression.


Analysis

For YARM, the high bandwidth YAUX laser loop with transfer function G ensures that the relative laser frequency fluctuations correlate with the relative length fluctuations as:

\frac{\delta \nu}{\nu} = \frac{G}{1 - G} {\frac{\delta L}{L}}

Then, getting the magnitude of the YARM displacement at calibration frequencies is possible by knowing the arm cavity length, open loop gain, and absolute frequency (wavelength). The relative calibration error on the magnitude of the displacement is

\frac{\Delta \lvert \delta L\lvert}{\lvert \delta L \lvert} = \left[ \left(\frac{\Delta {L}}{L}\right)^2+ \left(\frac{\Delta {\lambda}}{\lambda}\right)^2 + \left(\frac{\Delta {{\delta \nu}}}{{\delta \nu}}\right)^2 + \left(\frac{\Delta \lvert G \lvert}{\lvert G \lvert(\lvert G \lvert - 1)}\right)^2 \right]^{1/2}

including the relative uncertainties in the YARM length, wavelength, and open loop gain. Interestingly, the loop gain term weighs proportionally less as G increases, so even if G = 100 (10), its relative error contribution would be < 1%. To estimate our total error, we assume the wavelength and YARM length are 1064.1(5) and 37.79(1), and add the frequency dependent values for G with 10% error. Finally, we use the rms ASD to estimate the relative error from the beatnote fluctuation measurement.

The measurement was done similar to other instances, taking the 'C1:ALS-BEATY_FINE_PHASE_OUT_HZ_DQ' timeseries (sampled at 16 kHz) and demodulating at the calibration frequencies above to get the mean YAUX laser frequency fluctuation and its uncertainty from the demodulated rms ASD.


Results

Attachment #1 shows the raw timeseries, Attachment #2 shows the spectra around the cal lines, Attachment #3 shows the demodulated timeseries, Attachment #4 shows the final result for the 5 lines, including the tallied errors as detailed above.

ITMY actuation = 4.92(11) nm / count / f^2


Discussion

We compared our results from Attachment #4 against a MICH referenced ITMY actuation calibration found here; which Yuta guess-timated a 10% uncertainty (gray shaded band in Attachment #4). An important correction came for the 575 Hz line, not just because the YAUX OLG is small but because a violin filter on ITMY LSC output has a 1.4475 gain bump. In fact we collected any additional digital gains from the ITMY output filters:

Output filter digital gains at cal lines (from foton)
30.92 Hz 211.1 Hz 313.31 Hz 315.31 Hz 575.17 Hz
1.00007 1.0034 1.0101 1.01017 1.4475

Next

  • Consider moving the 575 Hz line to avoid additional digital gain, but try to remain at high frequency.
    • Maybe here we can use the resonant gain MOKU filters that Radhika is designing.
  • Setup a live loop gain calibration to reduce the uncertainty for the high frequency cal line.
    • We can also just grab GTRY (transmission) as a proxy for optical gain and use for budgeting.
  • Work on setting up constraints for error mitigation based on allan deviation of the beatnote and PDH nonlinearity.
  • Move this to FPMI or some other lock configuration
  15857   Wed Mar 3 12:00:58 2021 Paco, AnchalHowToIMCMC_F ASD

[Paco, Anchal]

- Saved BURT backup in /users/anchal/BURTsnaps/
- Copied existing code for mode cleaner noise budget from /users/rana/mat/mc. Will work on this from home to convert it inot new pynb way.

Get baseline IMC measurements (passive):
- MC_F:
  - What is MC_F? Let's find out.
  - On MC_F Cal window titled 'C1IOO-MC_FREQ', we turned off ON/OFF and back on again.
  - Using diaggui, we measured ASD of MC_F channel in units of counts/rtHz.

[Rana, Paco]

- Using diaggui, measured ASD from a template (under /users/Templates) and overlay the 1/f noise of the NPRO (Attachment 1)

[Anchal, Paco]

- WFS Master
  - Went through the schematic and tried to understand what is happening.
  - Accidentally switched on MC WF relief (python 3). Bunch of things were displayed on a terminal for a while and then we Ctrl-C it.
  - The only thing we noticed that change is a slight increase in WFS1 Yaw, and a corresponding decrease in WFS1 Pitch, WFS2 Pitch, and WFS2 Yaw.
  - We need to find out what this script does.


Future work:

  • Create an automated script for taking MC_F_DQ spectrum and refer it against reference trace.
  • Use pynb to create a noise budget for mode cleaner.
  • Identify excess noise between 10-40 Hz.
  • Configure output matrix in WFS Master to reduce the noise. Automate this process as well.
  15861   Thu Mar 4 10:54:12 2021 Paco, AnchalSummaryLSCPOY11 measurement, tried to lock Green Yend laser

[Paco, Anchal]

- First ran burtgooey as last time.

- Installed pyepics on base environment of donatella

ASS XARM:
- Clicked on ON in the drop down of "! More Scripts" below "! Scripts XARM" in C1ASS.adl
- Clicked on "Freeze Outputs" in the same menu after some time.
- Noticed that the sensing and output matrix of ASS on XARM and YARM look very different. The reason probably is because the YARM outputs have 4 TT1/2 P/Y dof instead of BS P/Y on the XARM. What are these TT1/2?

(Probably, unrelated but MC Unlocked and kept on trying to lock for about 10 minutes attaining the lock eventually.)

Locking XARM:
- From scripts/XARM we ran lockXarm.py from outside any conda environment using python command.
- Weirdly, we see that YARM is locked??? But XARM is not. Maybe this script is old.
- C1:LSC-TRY-OUTPUT went to around 0.75 (units unknown) while C1:LSC-TRX-OUTPUT is fluctuating around 0 only.

POY11 Spectrum measurement when YARM is locked:
- Created our own template as we couldn't find an existing one in users/Templates.
- Template file and data in Attachment 2.
- It is interesting to see most of the noise is in I quadrature with most noise in 10 to 100 Hz.
- Given the ARM is supposed to be much calmer than MC, this noise should be mostly due to the mode cleaner noise.
- We are not sure what units C1:LSC-POY11_I_ERR_DQ have, so Y scale is shown with out units.


Trying to lock Green YEND laser to YARM:
- We opened the Green Y shutter.
- We ensured that when temperature slider og green Y is moved up, the beatnote goes up.
- ARM was POY locked from previous step.
- Ran script scripts/YARM/Lock_ALS_YARM.py from outside any conda environment using python command.
- This locked green laser but unlocked the YARM POY.

Things moving around:
- Last step must have made all the suspension controls unstable.
- We see PRM and SRM QPDs moving a lot.
- Then we did burt restore to /opt/rtcds/caltech/c1/burt/autoburt/today/08:19/*.snap to go back to the state before we started changing things today.

[Paco left for vaccine appointment]

- However the unstable state didn't change from restore. I see a lot of movement in ITMX/Y. PRM and BS also now. Movement in WFS1 and MC2T as well.
 - I closed PSL shutter as well to hopefully disengage any loops that are still running unstably.
 - But at this point, it seems that the optics are just oscillating and need time to come back to rest. Hopefully we din't cause too much harm today :(.
 


My guess on what happened:

  • Us using the Lock_ALS_YARM.py probably created an unstable configuration in LSC matrix and was the start of the issue.
  • On seeing PRM fluctuate so much, we thought we should just burst restore everything. But that was a hammer to the problem.
  • This hammer probably changed the suspension position values suddenly causing an impulse to all the optics. So everything started oscillating.
  • Now MC WFS is waiting for MC to lock before it stablizes the mode cleaner. But MC autolocker is unable to lock because the optics are oscillating. Chicken-egg issue.
  • I'm not aware of how manually one can restore the state now. My only known guess is that if we wait for few hours, everything should calm back enough that MC can be locked and WFS servo can be switched on.
  15862   Thu Mar 4 11:59:25 2021 Paco, AnchalSummaryLSCWatchdog tripped, Optics damped back

Gautam came in and noted that the optics damping watchdogs had been tripped by a >5 magnitude earthquake somewhere off the coast of Australia. So, under guided assistance, we manually damped the optics using following:

  • Using the scripts/SUS/reEnableWatchdogs.py script we re-enabled all the watchdogs.
  • Everything except SRM was restored to stable state.
  • Then we clicked on SRM in SUS-> Watchdogs, disabled the Oplevs, shutdown the watchdog.
  • We changed the threshold for watchdog temporarily to 1000 to allow damping.
  • We enabled all the coil outputs  manually. Then enabled watchdog by clicking on Normal.
  • Once the SRM was damped, we shutdown the watchdog, brought back the threshold to 215 and restarted it.

Gautum also noticed that MC autolocker got turned OFF by me (Anchal), we turned it back on and MC engaged the lock again. All good, no harm done.

  15877   Mon Mar 8 12:01:02 2021 Paco, AnchalSummarytrainingInvestigate how-to XARM locking

[Paco, Anchal]

- Started zoom stream; thanks to whoever installed it!
- Spent some time trying to understand how anything we did last thursday lead to the sensing matrix change, but still cannot figure it out. 
- Tracking back on our actions, at ~10:30 we ran burt Restore with the 08:19/.*snap and in lack of a better suspect, we blame it on that action for now.

# ARM locking??
- Reading (not running) the scripts/XARM/lockXarm.py script and try to understand the workflow. It is pretty confusing that the result was to lock Yarm last time.
- It looks like this script was a copy of lockYarm.py, and was never updated (there's a chance we ran it for the first time last thursday)
- *Is there a script to lock the Arms?* Or should we write one? To write one, we first attempt a manual procedure;
    1. No need to change RFPD InMTRX
    2. All filters inputs / outputs are enabled 
    3. Outputs from XARM and YARM in the Output matrix are already going to ETMX and ETMY
      - Maybe we can have the ARM lock engage by changing the MC directly?
    4. Change C1:SUS-MC2_POS_OFFSET from -38 to -0, and enable C1:SUS-MC2_POS_OFFSET_ON
    5. Manually scan MC2_POS_OFFSET to 250 (nothing happens), then -250, then back to -38 (WFS1 PIT and YAW changed a little, but then returned to their nominal values)
      - Or maybe we need to provide the right gain...
    6. Disabled C1:SUS-MC2_POS_OFFSET_ON (back to nominal state)
    7. Look into manually changing C1:LSC-XARM_GAIN;
      From the command line using python:
      >> import epics
      >> ch_name = 'C1:LSC-XARM_GAIN'
      >> epics.caput(ch_name, 0.155) # nominal = 0.150
      - Could be unrelated, but we noted a slow spike on C1:PSL-FSS_PCDRIVE (definitely from before we changed anything)
      - Still nothing is happening
    8. Changed the gain to 0.175, then back to 0.150, no effect... then 0.2, 0.3 ...
      - Stop and check SUS_Watchdogs (should not have changed?) and everything remains nominal
      - Revert all changes symmetrically.
      - Could we have missed enabling FM1?
      - Briefly lost MC lock, but it came back on its own (probably unrelated)

- Wrap it up for the day. In summary; no harm done to our knowledge.

  15884   Tue Mar 9 10:57:06 2021 Paco, AnchalSummaryIMCXARM lock and POX spectra

[Paco, Anchal]

- Upon arrival, MC is locked, and we can see light in MON5 (PRM) (usually dark).

# XARM locking
- Read through "XARM POX" script (path='/cvs/cds/rtcds/caltech/c1/burt/c1configure/c1configureXarm')
- Before running the script, we noticed the PRM watchdog is down, so we manually repeat the procedure from last time, but see more swinging even though the watchdog is active.
- Run a reEnablePRMWatchdogs.py script (a copy of reEnableWatchdogs.py with optics=['PRM']), which had the same effect. 
- We manually disable the watchdog to recover the state we first encountered, and wait for the beam in MON5 to come to rest.
    - The question is; is it fine to lock Xarm with PRM watchdog down?
    - To investigate this, we look at the effect of the offset on the unwatchdog-PRM.
    - Manually change 'PRM_POS_OFFSET' to 200, and -800 (which is the value used in the script) with no effect on the PRM swinging.
- Moving on, run IFO > CONFIGURE > ! (X Arm) > RESTORE XARM (XARM POX), and ... success.

# MC-POX noise spectra
- With XARM locked, open diaggui and take spectra for C1:LSC-POX11_I_ERR_DQ, C1:LSC-POX11_Q_ERR_DQ, C1:IOO-MC_F_DQ
- Lost XARM lock while we were figuring out unit conversions...
    - Assuming 2.631e-13 m/counts (6941) and using 37.79 m (arm length), 1064.1 nm wavelength, we get a calibration factor of 2.631e-13 * c / (2*L*lambda) ~ 0.9809 Hz/count 
    - (FAQ?, how to find/compute/measure the correct calibration factors?)
- Relock XARM, retake spectra. Attachment 1 has plots for POX11_I/Q_ERR_DQ spectrum (cts/rtHz, we couldn't find relevant calibration) and MC_F_DQ in (Hz/rtHz from referring to 15576, we couldn't get the units to show on y scale.)

# MC-POY noise spectra (attempt)
- Now, run IFO > CONFIGURE > ! (Y Arm) > RESTORE YARM (YARM POY), and XARM locks (why?)
    - Could PRM watchdog being down be the cause? 
- Try C1ASS > (YARM) ! More Scripts > ON, and looked at YARM PIT/YAW striptool. 
- C1ASS > (YARM) ! Freeze Outputs, then OFF
- Go back to IFO > CONFIGURE > ! (Y Arm) > Align YARM  (ASS ON: Unfreeze), try running this then Freeze, then OFF Zero Outputs.
- Try RESTORE YARM (POY) again, still not working.
- Try RESTORE YARM ALS, then try again after opening the shutter, but also fail to lock AUX.
    - Is the PRM WD behind some evil misalignment? Will move forward with XARM bc it is happy.

# ARM locking
- Attempted the IFO > CONFIGURE > ! (X Arm) > RESTORE Xarm (XARM ALS) but green failed to lock and we lost XARM lock.
- Try to recover XARM lock... success. It's nice to have a (repeatable) checkpoint.
- Attempt YARM lock. Not successful. It just seems like the lock Triggers are not raised (misalignment?)
    - From C1SUS_ETMY, try changing the bias "C1:SUS-ETMY_YAW_OFFSET" manually to reduce the OPLEV_YERROR. Changed from -47 to -57.
    - Retry YARM lock script... no luck
    - From C1SUS_PRM, try changing the bias "C1:SUS-PRM_PIT_OFFSET" manually to reduce OPLEV errors. Changed from 34 to 22 with no effect, then realized the coil outputs are disabled because the WD is down...
    - So we do the following BIAS changes "C1:SUS-PRM_PIT_OFFSET" = 34 > 770 and "C1:SUS-PRM_YAW_OFFSET" = 134 > -6
    - Enable all Coil Outputs, turn WD to Normal, turn OPLEVs ON, (this time the beam does not swing like crazy).
    - Fine tune BIASes "C1:SUS-PRM_PIT_OFFSET" = 770 > 805  and "C1:SUS-PRM_YAW_OFFSET" = -6 > 65
        - Saw YARM locking briefly, then unlocking, but we stopped once the OPLEV_ERRs no longer overloaded (from magnitudes > 50 to ~ 40).
- Retry YARM lock... no luck
    - From C1SUS_ETMY, try changing the bias "C1:SUS-ETMY_PIT_OFFSET" from -1 to 6. 

Stop for the day. Leave XARM locked, MC locked. 

  15893   Wed Mar 10 11:46:22 2021 Paco, AnchalSummaryIMCIMC free swinging prep

[Paco, Anchal]

# Initial State
- MC is locked. The PRM monitor shows some oscillations.
- POP monitor shows light flashing once in a while.
- AS monitor shows one beam along with some other flashing beam around it.
- PRM Watchdog is tripped and shutdown. Everything else is normal except for overload on SRM OpLevs.
- Donatella got a mouse promotion

# Reenabling PRM watchdog:
- The custom reEnablePRMWatchdog.py has been deleted.
- Tried enabling the coil outputs manually and switching watchdog to Normal.
- Again saw large fluctuations like yesterday.
- Probably still the same issue of how current calculated actuations to the coils is in range -600 to -900 and gives and impulse to the optics when suddenly turned on.
- Waiting for PRM to damp down a little.
- Today we plan to change the position bias on PRM C1:SUS-PRM_POS_OFFSET instead of changing biases in pitch and yaw.
- Changing C1:SUS-PRM_POS_OFFSET from 0 to +/- 100 without enabling the coils, it seems upper and lower coils are anticorrelated with just changing the position. So going back to changing pitch.
- Changing C1:SUS-PRM_PIT_OFFSET from 0 -> 780. Switched on watchdog to normal.
- PRM damped down. OpLev errors are also within range.
- Enabled both OpLevs.

# Try locking Y-Arm
- IFO>CONFIGURE>YARM>Restore YARM (POY) using Donatella. See a bunch of python error messages in the call complaining about unable to find some python 2 files. Closed it with Ctrl-C after a stuck state.
- Tried running it on Pianosa, the script ran without error but Y-Arm didn't lock.

# Try locking X-Arm
- IFO>CONFIGURE>XARM>Restore XARM (POX) on Donatella. Again a bunch of OSError messages. Donatella is not configured properly to run scripts.
- Tried running it on Piasnosa, the script ran without error but X-Arm didn't lock.
- This might mean that both arms are misaligned or the BS/PRM is misaligned.
- Moving around C1:SUS-PRM_PIT_OFFSET and C1:SUS-PRM_YAW_OFFSET in order to see if the transmitted light is misalgined. Both arms are set to acquire lock if possible. No luck.

# Hypothesis: The Arm cavity is not aligned within itself (ITM-ETM)
- Will try to lock X-Arm with green light while tuning the ETMX. Hopefully the BS and ITM are aligned so that once we align ETMX to get a green lock, the IR will also lock from the other side.
- Running IFO>CONFIGURE>XARM>Restore XARM (ALS) on Pianosa. No lock, moving forward with tunning ETMX pitch and yaw offsets. Nothing changed. Brought back to same values.

[Rana joined, Anchal moved to Rossa from Pianosa]

# Moving on to IMC suspensions characterization:
- Closed the PSL shutter, to our suprise, the MC was still locked. We thought this would take away any light from IMC but it doesn't. Maybe the IFO Overview needs to show the schematic in a way where this doesn't happen: "No light from any laser entering the MC but it still is locked with a resonating field inside."
- Shutting IMCR shutter (hoping that would unlock the IMC), still nothing happend.
- Tried shutting PSL shutter from Rossa, nothing happened to MC lock still.
- Closed shutter IOO>Lock MC> Close PSL and this unlocked the IMC. Found out that this shutter channel is C1:PSL-PSL_ShutterRqst while the one from the sitemap>Shutter>PSL changes C1:AUX-PSL_ShutterRqst. Some clarification on these medm screens would be nice.
- Disabled the MC autolocked from IOO>Lock MC screen (C1:IOO-MC_LOCK_ENABLE).
- Checked the scripts/SUS/freeswing.py to understand how kick is delivered and optic is left to swing freely.
- Next, we are looking at the C1SUS_MC1 screen to understand what channels to read during data acquisition.
- In sensor matrix, we see INMON for each sensor which is probably raw counts data from the OSEMs. Rana mentioned that OSEM data comes out in units of microns. These are C1:SUS-MC1_ULSEN_OUTPUT (and so on for UR, LL, LR, SD).

- In prep for finishing, recovered Autolocker by first opening the PSL mechanical shutter, then re-enabling the Autolocker. The IMC lock didn't immediately recover, and we saw some fuzz on the PSL-FSS_FAST trace, so we closed the shutter again, waited a minute, then re-opened it and MC caught its lock.
 

  15897   Wed Mar 10 15:35:25 2021 Paco, AnchalSummaryIMCIMC free swinging experiment set to trigger at 5:00 am

A tmux session named "MCFreeSwingTest" will run on Rossa. This session is running script scripts/SUS/freeSwingMC.py (also attached) which will trigger at 5:00 am to impart 30000 counts kick to MC1, MC2, and MC3 after shutting PSL shutter and disabling the MC autolocker. It will let them freely swing for 1050 sec and will repeat 15 times to allow some averaging. In the end, it will undo all the changes it does and switches on autolocker on IMC. The script is set to restore any changes in case it fails at any point or a Ctrl-C is detected.

  15902   Thu Mar 11 08:13:24 2021 Paco, AnchalUpdateSUSIMC First Free Swing Test failed due to typo, restarting now

[Paco, Anchal]

The triggered code went on at 5:00 am today but a last minute change I made yesterday to increase number of repititions had an error and caused the script to exit putting everything back to normal. So as we came in the morning, we found the mode cleaner locked continuously after one free swing attempt at 5:00 am. I've fixed the script and ran it for 2 hours starting at 8;10 am. Our plan is to get some data atleast to play with when we are here. If the duration is not long enough, we'll try to run this again tomorrow morning. The new script is running on same tmux session 'MCFreeSwingTest' on Rossa

10:13 the script finished and IMC recovered lock.

Thu Mar 11 10:58:27 2021

The test ran succefully with the mode cleaner optics coming back to normal in the end of it. We wrote some scripts to read data and analyze it. More will come in future posts. No other changes were made today to the systems.

  15912   Fri Mar 12 11:44:53 2021 Paco, AnchalUpdatetrainingIMC SUS diagonalization in progress

[Paco, Anchal]

- Today we spent the morning shift debugging SUS input matrix diagonalization. MC stayed locked for most of the 4 hours we were here, and we didn't really touch any controls.